TI01222D/06/EN/04.18
71402584
2018-06-01
Products |
Solutions |
Services |
|
|
|
Technical Information
Proline Promass F 500
Coriolis flowmeter
The flowmeter offering premium accuracy and robustness, available as remote version with up to 4 I/Os
Application
•Measuring principle operates independently of physical fluid properties such as viscosity or density
•Highest measurement performance for liquids and gases under varying, demanding process conditions
Device properties
•Mass flow: measured error ±0.05 % (PremiumCal)
•Medium temperature: –196 to +350 °C (–320 to +662 °F)
•Nominal diameter: DN 8 to 250 (³⁄ to 10")
•Remote version with up to 4 I/Os
•Backlit display with touch control and WLAN access
•Standard cable between sensor and transmitter
Your benefits
•Highest process safety – immune to fluctuating and harsh environments
•Fewer process measuring points – multivariable measurement (flow, density, temperature)
•Space saving installation – no inlet/outlet run needs
•Full access to process and diagnostic information – numerous, freely combinable I/Os and fieldbuses
•Reduced complexity and variety – freely configurable I/O functionality
•Integrated verification – Heartbeat Technology
Proline Promass F 500
About this document . . . . . . . . . . . . . . . . . . . . . . . . 4
Symbols used . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . 4
Function and system design . . . . . . . . . . . . . . . . . . . 5
Measuring principle . . . . . . . . . . . . . . . . . . . . . . . . . . . . 5 Measuring system . . . . . . . . . . . . . . . . . . . . . . . . . . . . . 7 Equipment architecture . . . . . . . . . . . . . . . . . . . . . . . . . 9 Safety . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . 9
Input . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . 12
Measured variable . . . . . . . . . . . . . . . . . . . . . . . . . . . . 12 Measuring range . . . . . . . . . . . . . . . . . . . . . . . . . . . . . 12 Operable flow range . . . . . . . . . . . . . . . . . . . . . . . . . . . 13 Input signal . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . 13
Output . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . 15
Output and input variants . . . . . . . . . . . . . . . . . . . . . . . 15 Output signal . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . 16 Signal on alarm . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . 20 Ex connection data . . . . . . . . . . . . . . . . . . . . . . . . . . . 22 Low flow cut off . . . . . . . . . . . . . . . . . . . . . . . . . . . . . 23 Galvanic isolation . . . . . . . . . . . . . . . . . . . . . . . . . . . . 23 Protocol-specific data . . . . . . . . . . . . . . . . . . . . . . . . . . 24
Power supply . . . . . . . . . . . . . . . . . . . . . . . . . . . . . |
28 |
Terminal assignment . . . . . . . . . . . . . . . . . . . . . . . . . . |
28 |
Device plugs available . . . . . . . . . . . . . . . . . . . . . . . . . . |
29 |
Pin assignment, device plug . . . . . . . . . . . . . . . . . . . . . . |
30 |
Supply voltage . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . |
32 |
Power consumption . . . . . . . . . . . . . . . . . . . . . . . . . . . |
32 |
Current consumption . . . . . . . . . . . . . . . . . . . . . . . . . . |
32 |
Power supply failure . . . . . . . . . . . . . . . . . . . . . . . . . . |
32 |
Electrical connection . . . . . . . . . . . . . . . . . . . . . . . . . . |
33 |
Potential equalization . . . . . . . . . . . . . . . . . . . . . . . . . |
43 |
terminals . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . |
44 |
Cable entries . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . |
44 |
Cable specification . . . . . . . . . . . . . . . . . . . . . . . . . . . . |
44 |
Performance characteristics . . . . . . . . . . . . . . . . . . 49
Reference operating conditions . . . . . . . . . . . . . . . . . . . 49 Maximum measured error . . . . . . . . . . . . . . . . . . . . . . . 49 Repeatability . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . 51 Response time . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . 52 Influence of ambient temperature . . . . . . . . . . . . . . . . . 52 Influence of medium temperature . . . . . . . . . . . . . . . . . . 52 Influence of medium pressure . . . . . . . . . . . . . . . . . . . . 53 Design fundamentals . . . . . . . . . . . . . . . . . . . . . . . . . . 53
Installation . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . 54
Mounting location . . . . . . . . . . . . . . . . . . . . . . . . . . . . 54 Orientation . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . 55 Inlet and outlet runs . . . . . . . . . . . . . . . . . . . . . . . . . . 56 Mounting the transmitter housing . . . . . . . . . . . . . . . . . 57 Special mounting instructions . . . . . . . . . . . . . . . . . . . . 58
Environment . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . 60
Ambient temperature range . . . . . . . . . . . . . . . . . . . . . 60
Storage temperature . . . . . . . . . . . . . . . . . . . . . . . . . . 60 Climate class . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . 60 Degree of protection . . . . . . . . . . . . . . . . . . . . . . . . . . 60 Vibration resistance . . . . . . . . . . . . . . . . . . . . . . . . . . . 60 Shock resistance . . . . . . . . . . . . . . . . . . . . . . . . . . . . . 61 Shock resistance . . . . . . . . . . . . . . . . . . . . . . . . . . . . . 61 Interior cleaning . . . . . . . . . . . . . . . . . . . . . . . . . . . . . 61 Electromagnetic compatibility (EMC) . . . . . . . . . . . . . . . 61
Process . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . 61
Medium temperature range . . . . . . . . . . . . . . . . . . . . . . 61 Density . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . 62 Pressure-temperature curves . . . . . . . . . . . . . . . . . . . . . 63 Sensor housing . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . 68 Rupture disk . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . 69 Flow limit . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . 69 Pressure loss . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . 69 System pressure . . . . . . . . . . . . . . . . . . . . . . . . . . . . . 69 Thermal insulation . . . . . . . . . . . . . . . . . . . . . . . . . . . 70 Heating . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . 70 Vibrations . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . 71
Custody transfer measurement . . . . . . . . . . . . . . . 72
Mechanical construction . . . . . . . . . . . . . . . . . . . . 73
Dimensions in SI units . . . . . . . . . . . . . . . . . . . . . . . . . 73 Dimensions in US units . . . . . . . . . . . . . . . . . . . . . . . . . 96 Weight . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . 108 Materials . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . 109 Process connections . . . . . . . . . . . . . . . . . . . . . . . . . . 112 Surface roughness . . . . . . . . . . . . . . . . . . . . . . . . . . . 112
Operability . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . 112
Operating concept . . . . . . . . . . . . . . . . . . . . . . . . . . . 112 Languages . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . 113 Local operation . . . . . . . . . . . . . . . . . . . . . . . . . . . . . 113 Remote operation . . . . . . . . . . . . . . . . . . . . . . . . . . . 113 Service interface . . . . . . . . . . . . . . . . . . . . . . . . . . . . 119 Network integration . . . . . . . . . . . . . . . . . . . . . . . . . . 120 Supported operating tools . . . . . . . . . . . . . . . . . . . . . . 121 HistoROM data management . . . . . . . . . . . . . . . . . . . . 122
Certificates and approvals . . . . . . . . . . . . . . . . . . 124
CE mark . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . 124 C-Tick symbol . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . 124 Ex approval . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . 124 Sanitary compatibility . . . . . . . . . . . . . . . . . . . . . . . . . 127 Pharmaceutical compatibility . . . . . . . . . . . . . . . . . . . . 127 Functional safety . . . . . . . . . . . . . . . . . . . . . . . . . . . . 127 HART certification . . . . . . . . . . . . . . . . . . . . . . . . . . . 127 FOUNDATION Fieldbus certification . . . . . . . . . . . . . . . 127 Certification PROFIBUS . . . . . . . . . . . . . . . . . . . . . . . . 127 EtherNet/IP certification . . . . . . . . . . . . . . . . . . . . . . . 128 Certification PROFINET . . . . . . . . . . . . . . . . . . . . . . . . 128 Pressure Equipment Directive . . . . . . . . . . . . . . . . . . . 128 Radio approval . . . . . . . . . . . . . . . . . . . . . . . . . . . . . 128 Measuring instrument approval . . . . . . . . . . . . . . . . . . 128 Additional certification . . . . . . . . . . . . . . . . . . . . . . . . 128
2 |
Endress+Hauser |
Proline Promass F 500
Other standards and guidelines . . . . . . . . . . . . . . . . . . 129
Ordering information . . . . . . . . . . . . . . . . . . . . . . 130
Application packages . . . . . . . . . . . . . . . . . . . . . . 130
Diagnostics functions . . . . . . . . . . . . . . . . . . . . . . . . . 130 Heartbeat Technology . . . . . . . . . . . . . . . . . . . . . . . . 130 Concentration . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . 131 Special density . . . . . . . . . . . . . . . . . . . . . . . . . . . . . 131 Petroleum . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . 131 OPC-UA server . . . . . . . . . . . . . . . . . . . . . . . . . . . . . 131
Accessories . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . 131
Device-specific accessories . . . . . . . . . . . . . . . . . . . . . . 132 Communication-specific accessories . . . . . . . . . . . . . . . 133 Service-specific accessories . . . . . . . . . . . . . . . . . . . . . 134 System components . . . . . . . . . . . . . . . . . . . . . . . . . . 134
Supplementary documentation . . . . . . . . . . . . . . 135
Standard documentation . . . . . . . . . . . . . . . . . . . . . . . 135 Device-dependent additional documentation . . . . . . . . . 136
Registered trademarks . . . . . . . . . . . . . . . . . . . . . 136
Endress+Hauser |
3 |
Proline Promass F 500
Symbols used |
Electrical symbols |
|
||
|
|
|
|
|
|
Symbol |
Meaning |
||
|
|
|
|
|
|
|
|
|
Direct current |
|
|
|
|
|
|
|
|
|
Alternating current |
|
|
|
|
|
|
|
|
|
Direct current and alternating current |
|
|
|
|
|
|
|
|
|
|
|
|
|
|
Ground connection |
|
|
|
|
A grounded terminal which, as far as the operator is concerned, is grounded via a |
|
|
|
|
grounding system. |
|
|
|
|
|
|
|
|
|
Protective Earth (PE) |
|
|
|
|
A terminal which must be connected to ground prior to establishing any other |
|
|
|
|
connections. |
|
|
|
|
The ground terminals are situated inside and outside the device: |
|
|
|
|
• Inner ground terminal: Connects the protectiv earth to the mains supply. |
|
|
|
|
• Outer ground terminal: Connects the device to the plant grounding system. |
|
|
|
|
|
Communication symbols
Symbol Meaning
Wireless Local Area Network (WLAN)
Communication via a wireless, local network.
LED
Light emitting diode is off.
LED
Light emitting diode is on.
LED
Light emitting diode is flashing.
Symbols for certain types of information
Symbol Meaning
Permitted
Procedures, processes or actions that are permitted.
Preferred
Procedures, processes or actions that are preferred.
Forbidden
Procedures, processes or actions that are forbidden.
Tip
Indicates additional information.
Reference to documentation.
A |
Reference to page. |
Reference to graphic.
Visual inspection.
4 |
Endress+Hauser |
Proline Promass F 500
Symbols in graphics
|
Symbol |
Meaning |
|||||
|
|
|
|
|
|
|
|
1, 2, 3, ... |
Item numbers |
||||||
|
|
|
|
|
|
|
|
|
|
, |
|
, |
|
, … |
Series of steps |
|
1. |
2. |
3. |
||||
|
|
|
|||||
|
A, B, C, ... |
Views |
|||||
|
|
||||||
A-A, B-B, C-C, ... |
Sections |
-Hazardous area
. |
Safe area (non-hazardous area) |
|
|
|
|
|
Flow direction |
Measuring principle |
The measuring principle is based on the controlled generation of Coriolis forces. These forces are |
|
|
always present in a system when both translational and rotational movements are superimposed. |
|
|
Fc = |
2 · ∆m (ν · ω) |
|
Fc = |
Coriolis force |
|
∆m = |
moving mass |
|
ω = |
rotational velocity |
|
ν = |
radial velocity in rotating or oscillating system |
The amplitude of the Coriolis force depends on the moving mass ∆m, its velocity ν in the system and thus on the mass flow. Instead of a constant rotational velocity ω, the sensor uses oscillation.
In the sensor, two parallel measuring tubes containing flowing fluid oscillate in antiphase, acting like a tuning fork. The Coriolis forces produced at the measuring tubes cause a phase shift in the tube oscillations (see illustration):
•At zero flow (when the fluid is at a standstill) the two tubes oscillate in phase (1).
•Mass flow causes deceleration of the oscillation at the inlet of the tubes (2) and acceleration at the outlet (3).
|
|
|
|
|
|
1 |
2 |
3 |
|
A0028850
The phase difference (A-B) increases with increasing mass flow. Electrodynamic sensors register the tube oscillations at the inlet and outlet. System balance is ensured by the antiphase oscillation of the two measuring tubes. The measuring principle operates independently of temperature, pressure, viscosity, conductivity and flow profile.
Endress+Hauser |
5 |
Proline Promass F 500
Density measurement
The measuring tube is continuously excited at its resonance frequency. A change in the mass and thus the density of the oscillating system (comprising measuring tube and fluid) results in a corresponding, automatic adjustment in the oscillation frequency. Resonance frequency is thus a function of medium density. The microprocessor utilizes this relationship to obtain a density signal.
Volume measurement
Together with the measured mass flow, this is used to calculate the volume flow.
Temperature measurement
The temperature of the measuring tube is determined in order to calculate the compensation factor due to temperature effects. This signal corresponds to the process temperature and is also available as an output signal.
6 |
Endress+Hauser |
Proline Promass F 500
Measuring system |
The measuring system consists of a transmitter and a sensor. The transmitter and sensor are |
|
mounted in physically separate locations. They are interconnected by connecting cables. |
Transmitter
Two versions of the transmitter are available.
Proline 500 – digital |
Proline 500 |
||||
|
|
|
|
|
|
For use in applications not required to meet special requirements due to |
For use in applications required to meet special requirements due to |
||||
ambient or operating conditions. |
ambient or operating conditions. |
||||
|
|
|
|
|
|
|
|
|
|
|
|
1 |
1 |
|
A
B
2
2
3 |
3 |
A |
Non-hazardous area or Zone 2; Class I, Division 2 |
|
Non-hazardous area or Zone 2; Class I, Division 2 or Zone 1; Class I, |
B |
Non-hazardous area or Zone 2; Class I, Division 2 or Zone 1; Class I, |
|
Division 1 |
|
Division 1 |
1 |
Transmitter with integrated ISEM |
1 |
Transmitter |
2 |
Connecting cable: cable, separate |
2 |
Connecting cable: cable, separate, standard |
3 |
Sensor connection housing |
3Sensor connection housing with integrated ISEM
• |
Flexible and cost-effective separate installation. |
Application examples for sensors without electronics: |
• A standard cable can be used as the connecting cable. |
• Strong vibrations at the sensor. |
|
|
|
• Sensor in underground installations. |
|
|
• Permanent immersion of sensor in water, IP68 ingress protection. |
• Electronics in the transmitter housing, ISEM (intelligent sensor |
• Electronics and ISEM (intelligent sensor electronics module) in the |
|
|
electronics module) in the sensor connection housing |
transmitter housing |
• |
Signal transmission: digital |
• Signal transmission: analog |
|
Order code for "Integrated ISEM electronics", option A "Sensor" |
Order code for "Integrated ISEM electronics", option B "Transmitter" |
|
|
|
Connecting cable (can be ordered in various lengths)→ 132 |
|
|
|
|
|
• Length: |
• Length: max. 20 m (65 ft) |
|
|
– Zone 2; Class I, Division 2: max. 300 m (1 000 ft) |
• Cable with a common shield and individual shielded cores (3 pairs) |
– Zone 1; Class I, Division 1: max. 150 m (500 ft)
•Standard cable with common shield (pair-stranded)
Hazardous area
Use in: Zone 2; Class I, Division 2 |
Use in: Zone 1; Class I, Division 1 oder Zone 2; Class I, Division 2 |
Mixed installation is possible:
•Sensor: Zone 1; Class I, Division 1
•Transmitter: Zone 2; Class I, Division 2
Device versions and materials
• |
Transmitter housing |
• |
Transmitter housing |
||
|
– |
Aluminum, coated: aluminum, AlSi10Mg, coated |
|
– |
Aluminum, coated: aluminum, AlSi10Mg, coated |
|
– |
Material: polycarbonate |
|
– |
Cast, stainless: cast, stainless steel, 1.4409 (CF3M) similar to 316L |
• |
Material of window in transmitter housing |
• |
Window material: glass |
–Aluminum, coated: glass
–Polycarbonate: plastic
Configuration
•External operation via 4-line, backlit, graphic local display with touch control and guided menus ("Make-it-run" wizards) for application-specific commissioning.
•Via service interface or WLAN interface:
–Operating tools (e.g. FieldCare, DeviceCare, SmartBlue app)
–Web server (access via Web browser, e.g. Microsoft Internet Explorer, Microsoft Edge)
Endress+Hauser |
7 |
Proline Promass F 500
Sensor connection housing
Different versions of the connection housing are available.
Order code for "Sensor connection housing", option A, "Aluminum, coated": Aluminum, AlSi10Mg, coated
This device version is only available in conjunction with the Proline 500 – digital transmitter.
Order code for "Sensor connection housing", option B, "Stainless":
• Hygienic version, stainless steel 1.4301 (304)
• Optional: order code for "Sensor feature", option CC "Hygienic version, for maximum corrosion resistance": stainless steel 1.4404 (316L)
Order code for "Sensor connection housing", option C, "Ultra-compact hygienic, stainless":
• Hygienic version, stainless steel 1.4301 (304)
•Optional: order code for "Sensor feature", option CC "Hygienic version, for maximum corrosion resistance": stainless steel 1.4404 (316L)
This device version is only available in conjunction with the Proline 500 – digital transmitter.
Order code for "Sensor connection housing", option L, "Cast, stainless": 1.4409 (CF3M) similar to 316L
Sensor
PromassF |
|
• Excellent performance across a wide range of applications |
|
|
• Simultaneous measurement of flow, volume flow, density and |
|
|
temperature (multivariable) |
|
|
• Immune to process influences |
|
|
• Materials: |
|
|
– Sensor: stainless steel, 1.4301/1.4307 (304L); optional 1.4404 |
|
A0026713 |
(316/316L) |
–Measuring tubes: stainless steel, 1.4539 (904L); 1.4404 (316/316L); Alloy C22, 2.4602 (UNS N06022)
–Process connections: stainless steel, 1.4404 (316/316L); 1.4301 (304); Alloy C22, 2.4602 (UNS N06022)
8 |
Endress+Hauser |
Proline Promass F 500 |
|
Equipment architecture |
|
1 |
|
5 |
|
2 |
3 |
|
|
|
4 |
6 |
|
|
3 |
7 |
|
|
A0027512 |
1 Possibilities for integrating measuring devices into a system |
|
1Control system (e.g. PLC)
2Connecting cable (0/4 to 20 mA HART etc.)
3Fieldbus
4Segment coupler
5Non-hazardous area
6 Hazardous area: Zone 2; Class I, Division 2
7Hazardous area: Zone 1; Class I, Division 1
Safety |
IT security |
Our warranty is valid only if the device is installed and used as described in the Operating Instructions. The device is equipped with security mechanisms to protect it against any inadvertent changes to the settings.
IT security measures, which provide additional protection for the device and associated data transfer, must be implemented by the operators themselves in line with their security standards.
Device-specific IT security
The device offers a range of specific functions to support protective measures on the operator's side. These functions can be configured by the user and guarantee greater in-operation safety if used correctly. An overview of the most important functions is provided in the following section.
Function/interface |
Factory setting |
Recommendation |
|
|
|
Write protection via hardware write |
Not enabled. |
On an individual basis following risk |
protection switch → 10 |
|
assessment. |
|
|
|
Access code |
Not enabled |
Assign a customized access code during |
(also applies for Web server login or |
(0000). |
commissioning. |
FieldCare connection) → 10 |
|
|
|
|
|
WLAN |
Enabled. |
On an individual basis following risk |
(order option in display module) |
|
assessment. |
|
|
|
WLAN security mode |
Enabled (WPA2- |
Do not change. |
|
PSK) |
|
|
|
|
WLAN passphrase |
Serial number |
Assign a customized access code during |
(password) → 10 |
|
commissioning. |
|
|
|
WLAN mode |
Access Point |
On an individual basis following risk |
|
|
assessment. |
|
|
|
Web server→ 10 |
Enabled. |
On an individual basis following risk |
|
|
assessment. |
|
|
|
CDI-RJ45 service interface → 11 |
– |
On an individual basis following risk |
|
|
assessment. |
|
|
|
Endress+Hauser |
9 |
Proline Promass F 500
Protecting access via hardware write protection
Write access to the device parameters via the local display, Web browser or operating tool (e.g. FieldCare, DeviceCare) can be disabled via a write protection switch (DIP switch on the motherboard). When hardware write protection is enabled, only read access to the parameters is possible.
Hardware write protection is disabled when the device is delivered.
Protecting access via a password
Different passwords are available to protect write access to the device parameters or access to the device via the WLAN interface.
•User-specific access code
Protect write access to the device parameters via the local display, Web browser or operating tool (e.g. FieldCare, DeviceCare). Access authorization is clearly regulated through the use of a userspecific access code.
•WLAN passphrase
The network key protects a connection between an operating unit (e.g. notebook or tablet) and the device via the WLAN interface which can be ordered as an option.
•Infrastructure mode
When the device is operated in infrastructure mode, the WLAN passphrase corresponds to the WLAN passphrase configured on the operator side.
User-specific access code
Write access to the device parameters via the local display, Web browser or operating tool (e.g. FieldCare, DeviceCare) can be protected by the modifiable, user-specific access code.
WLAN passphrase: Operation as WLAN access point
A connection between an operating unit (e.g. notebook or tablet) and the device via the WLAN interface, which can be ordered as an optional extra, is protected by the network key. The WLAN authentication of the network key complies with the IEEE 802.11 standard.
When the device is delivered, the network key is pre-defined depending on the device. It can be changed via the WLAN settings submenu in the WLAN passphrase parameter.
Infrastructure mode
A connection between the device and WLAN access point is protected by means of an SSID and passphrase on the system side. Please contact the relevant system administrator for access.
General notes on the use of passwords
•The access code and network key supplied with the device should be changed during commissioning.
•Follow the general rules for generating a secure password when defining and managing the access code or network key.
•The user is responsible for the management and careful handling of the access code and network key.
Access via Web server
The device can be operated and configured via a Web browser with the integrated Web server. The connection is via the service interface (CDI-RJ45) or the WLAN interface. For device versions with the EtherNet/IP and PROFINET communication protocols, the connection can also be established via the terminal connection for signal transmission with EtherNet/IP or PROFINET (RJ45 connector).
The Web server is enabled when the device is delivered. The Web server can be disabled if necessary (e.g. after commissioning) via the Web server functionality parameter.
The device and status information can be hidden on the login page. This prevents unauthorized access to the information.
For detailed information on device parameters, see:
The "Description of Device Parameters" document → 135
Access via OPC-UA
The "OPC UA Server" application package is available in the device version with the HART communication protocol → 131.
10 |
Endress+Hauser |
Proline Promass F 500
The device can communicate with OPC UA clients using the "OPC UA Server" application package.
The OPC UA server integrated in the device can be accessed via the WLAN access point using the WLAN interface - which can be ordered as an optional extra - or the service interface (CDIRJ45) via Ethernet network. Access rights and authorization as per separate configuration.
The following Security Modes are supported as per the OPC UA Specification (IEC 62541):
•None
•Basic128Rsa15 – signed
•Basic128Rsa15 – signed and encrypted
Access via service interface (CDI-RJ45)
The device can be connected to a network via the service interface (CDI-RJ45). Device-specific functions guarantee the secure operation of the device in a network.
The use of relevant industrial standards and guidelines that have been defined by national and international safety committees, such as IEC/ISA62443 or the IEEE, is recommended. This includes organizational security measures such as the assignment of access authorization as well as technical measures such as network segmentation.
The device can be integrated in a ring topology. The device is integrated via the terminal connection for signal transmission (output 1) and the connection to the service interface (CDIRJ45) → 117.
Endress+Hauser |
11 |
Proline Promass F 500
|
Input |
|
|
|
|
|
|
|
|
|
|
Measured variable |
Direct measured variables |
|
|
|
|
|
• Mass flow |
|
|
|
|
|
• Density |
|
|
|
|
|
• Temperature |
|
|
|
|
|
Calculated measured variables |
|
|
|
|
|
• Volume flow |
|
|
|
|
|
• Corrected volume flow |
|
|
|
|
|
• Reference density |
|
|
|
|
|
|
|
|
|
|
Measuring range |
Measuring ranges for liquids |
|
|
|
|
|
|
|
|
|
|
|
DN |
|
Measuring range full scale values min(F) to max(F) |
||
|
[mm] |
|
[in] |
[kg/h] |
[lb/min] |
|
|
|
|
|
|
|
8 |
|
³⁄ |
0 to 2 000 |
0 to 73.50 |
|
|
|
|
|
|
|
15 |
|
½ |
0 to 6 500 |
0 to 238.9 |
|
|
|
|
|
|
|
25 |
|
1 |
0 to 18 000 |
0 to 661.5 |
|
|
|
|
|
|
|
40 |
|
1½ |
0 to 45 000 |
0 to 1 654 |
|
|
|
|
|
|
|
50 |
|
2 |
0 to 70 000 |
0 to 2 573 |
|
|
|
|
|
|
|
80 |
|
3 |
0 to 180 000 |
0 to 6 615 |
|
|
|
|
|
|
|
100 |
|
4 |
0 to 350 000 |
0 to 12 860 |
|
|
|
|
|
|
|
150 |
|
6 |
0 to 800 000 |
0 to 29 400 |
|
|
|
|
|
|
|
250 |
|
10 |
0 to 2 200 000 |
0 to 80 850 |
|
|
|
|
|
|
Measuring range for gases
The full scale value depends on the density and the sound velocity of the gas used and can be calculated with the formula below:
max(G) = minimum ( max(F) · ρG : x ; ρG · cG · π/2 · (di)2 · 3600)
max(G) |
Maximum full scale value for gas [kg/h] |
|
||
max(F) |
Maximum full scale value for liquid [kg/h] |
|
||
max(G) < max(F) |
max(G) can never be greater than max(F) |
|
||
ρG |
Gas density in [kg/m³] at operating conditions |
|
||
|
|
|
|
|
x |
Constant dependent on nominal diameter |
|
||
|
|
|
|
|
cG |
Sound velocity (gas) [m/s] |
|
||
di |
Measuring tube internal diameter [m] |
|
||
|
|
|
|
|
DN |
|
|
x |
|
|
|
|
|
|
[mm] |
|
[in] |
|
[kg/m3] |
8 |
|
³⁄ |
|
60 |
|
|
|
|
|
15 |
|
½ |
|
80 |
|
|
|
|
|
25 |
|
1 |
|
90 |
|
|
|
|
|
40 |
|
1½ |
|
90 |
|
|
|
|
|
50 |
|
2 |
|
90 |
|
|
|
|
|
80 |
|
3 |
|
110 |
|
|
|
|
|
12 |
Endress+Hauser |
Proline Promass F 500
|
DN |
x |
|
[mm] |
|
[in] |
[kg/m3] |
|
|||
100 |
|
4 |
130 |
|
|
|
|
150 |
|
6 |
200 |
|
|
|
|
250 |
|
10 |
200 |
|
|
|
|
To calculate the measuring range, use the Applicator sizing tool → 134
Calculation example for gas
•Sensor: Promass F, DN 50
•Gas: Air with a density of 60.3 kg/m³ (at 20 °C and 50 bar)
•Measuring range (liquid): 70 000 kg/h
•x = 90 kg/m³ (for Promass F, DN 50)
Maximum possible full scale value:
max(G) = max(F) · ρG : x = 70 000 kg/h · 60.3 kg/m³ : 90 kg/m³ = 46 900 kg/h
|
Recommended measuring range |
|
"Flow limit" section → 69 |
|
|
Operable flow range |
Over 1000 : 1. |
|
Flow rates above the preset full scale value do not override the electronics unit, with the result that |
|
the totalizer values are registered correctly. |
|
|
Input signal |
Input and output versions |
|
→ 15 |
|
External measured values |
|
To increase the accuracy of certain measured variables or to calculate the corrected volume flow for |
|
gases, the automation system can continuously write different measured values to the measuring |
|
device: |
|
• Operating pressure to increase accuracy (Endress+Hauser recommends the use of a pressure |
|
measuring device for absolute pressure, e.g. Cerabar M or Cerabar S) |
|
• Medium temperature to increase accuracy (e.g. iTEMP) |
|
• Reference density for calculating the corrected volume flow for gases |
|
Various pressure transmitters and temperature measuring devices can be ordered from Endress |
|
+Hauser: see "Accessories" section → 134 |
|
It is recommended to read in external measured values to calculate the corrected volume flow. |
|
HART protocol |
|
The measured values are written from the automation system to the measuring device via the HART |
|
protocol. The pressure transmitter must support the following protocol-specific functions: |
|
• HART protocol |
|
• Burst mode |
|
Current input |
|
The measured values are written from the automation system to the measuring device via the |
|
current input → 14. |
|
Digital communication |
|
The measured values can be written from the automation system to the measuring via: |
|
• FOUNDATION Fieldbus |
|
• PROFIBUS DP |
|
• PROFIBUS PA |
|
• Modbus RS485 |
|
• EtherNet/IP |
|
• PROFINET |
Endress+Hauser |
13 |
Proline Promass F 500
Current input 0/4 to 20 mA
Current input |
0/4 to 20 mA (active/passive) |
|
|
|
|
Current span |
• 4 to 20 mA (active) |
|
|
• 0/4 to 20 mA (passive) |
|
|
|
|
Resolution |
1 |
µA |
|
|
|
Voltage drop |
Typically: 0.6 to 2 V for 3.6 to 22 mA (passive) |
|
|
|
|
Maximum input voltage |
≤ 30 V (passive) |
|
|
|
|
Open-circuit voltage |
≤ 28.8 V (active) |
|
|
|
|
Possible input variables |
• Pressure |
|
|
• Temperature |
|
|
• Density |
|
|
|
|
Status input
Maximum input values |
• DC –3 to 30 V |
|
|
• If status input is active (ON): Ri >3 kΩ |
|
Response time |
Adjustable: 5 to 200 ms |
|
|
|
|
Input signal level |
• Low signal: DC –3 to +5 V |
|
|
• High signal: DC 12 to 30 V |
|
|
|
|
Assignable functions |
• Off |
|
|
• |
Reset the individual totalizers separately |
|
• |
Reset all totalizers |
|
• Flow override |
|
|
|
|
14 |
Endress+Hauser |
Proline Promass F 500
|
Output |
|
|
|
|
|
|
|
|
|
|
|
|
||||||||||
Output and input variants |
Depending on the option selected for output/input 1, different options are available for the other |
||||||||||
|
outputs and inputs. Only one option can be selected for each output/input 1 to 4. The table must be |
||||||||||
|
read vertically (↓). |
|
|
|
|
|
|
|
|
|
|
|
Example: If the option BA "4–20 mA HART" was selected for output/input 1, one of the options A, B, |
||||||||||
|
D, E, F, H, I or J is available for output 2, and one of the options A, B, D, E, F, H, I or J is available for |
||||||||||
|
output 3 and 4. |
|
|
|
|
|
|
|
|
|
|
|
|
|
|
|
|
|
|
|
|
|
|
Order code for "Output; input 1" (020) → |
|
|
|
|
Possible options |
|
|
|
|||
|
|
|
|
|
|
|
|
|
|
|
|
Current output 4 to 20 mA HART |
|
BA |
|
|
|
|
|
|
|
|
|
|
|
|
|
|
|
|
|
|
|
|
|
Current output 4 to 20 mA HART Ex i |
↓ |
CA |
|
|
|
|
|
|
|
|
|
|
|
|
|
|
|
|
|
|
|
|
|
FOUNDATION Fieldbus |
|
|
↓ |
SA |
|
|
|
|
|
|
|
|
|
|
|
|
|
|
|
|
|
|
|
FOUNDATION Fieldbus Ex i |
|
|
|
↓ |
TA |
|
|
|
|
|
|
|
|
|
|
|
|
|
|
|
|
|
|
PROFIBUS DP |
|
|
|
|
↓ |
LA |
|
|
|
|
|
|
|
|
|
|
|
|
|
|
|
|
|
PROFIBUS PA |
|
|
|
|
|
↓ |
GA |
|
|
|
|
|
|
|
|
|
|
|
|
|
|
|
|
PROFIBUS PA Ex i |
|
|
|
|
|
|
↓ |
HA |
|
|
|
|
|
|
|
|
|
|
|
|
|
|
|
Modbus RS485 |
|
|
|
|
|
|
|
↓ |
MA |
|
|
|
|
|
|
|
|
|
|
|
|
|
|
EtherNet/IP 2-port switch integrated |
|
|
|
|
|
|
|
|
↓ |
NA |
|
|
|
|
|
|
|
|
|
|
|
|
|
PROFINET 2-port switch integrated |
|
|
|
|
|
|
|
|
|
↓ |
RA |
|
|
|
|
|
|
|
|
|
|
|
|
Order code for "Output; input 2" (021) → |
↓ |
↓ |
↓ |
↓ |
↓ |
↓ |
↓ |
↓ |
↓ |
↓ |
|
|
|
|
|
|
|
|
|
|
|
|
|
Not assigned |
|
A |
A |
A |
A |
A |
A |
A |
A |
A |
A |
|
|
|
|
|
|
|
|
|
|
|
|
Current output 0/4 to 20 mA |
|
B |
|
B |
|
B |
B |
|
B |
B |
B |
|
|
|
|
|
|
|
|
|
|
|
|
Current output 0/4 to 20 mA (Ex i) |
|
|
C |
|
C |
|
|
C |
|
|
|
|
|
|
|
|
|
|
|
|
|
|
|
User configurable input/output 1) |
|
D |
|
D |
|
D |
D |
|
D |
D |
D |
Pulse/frequency/switch output |
|
E |
|
E |
|
E |
E |
|
E |
E |
E |
|
|
|
|
|
|
|
|
|
|
|
|
Double pulse output 2) |
|
F |
|
|
|
|
|
|
F |
|
|
Pulse/frequency/switch output (Ex i) |
|
|
G |
|
G |
|
|
G |
|
|
|
|
|
|
|
|
|
|
|
|
|
|
|
Relay output |
|
H |
|
H |
|
H |
H |
|
H |
H |
H |
|
|
|
|
|
|
|
|
|
|
|
|
Current input 0/4 to 20 mA |
|
I |
|
I |
|
I |
I |
|
I |
I |
I |
|
|
|
|
|
|
|
|
|
|
|
|
Status input |
|
J |
|
J |
|
J |
J |
|
J |
J |
J |
|
|
|
|
|
|
|
|
|
|
|
|
Order code for "Output; input 3" (022), "Output; input 4" (023) 3) → |
↓ |
↓ |
↓ |
↓ |
↓ |
↓ |
↓ |
↓ |
↓ |
↓ |
|
Not assigned |
|
A |
A |
A |
A |
A |
A |
A |
A |
A |
A |
|
|
|
|
|
|
|
|
|
|
|
|
Current output 0/4 to 20 mA |
|
B |
|
|
|
B |
|
|
B |
B |
B |
|
|
|
|
|
|
|
|
|
|
|
|
Current output 0/4 to 20 mA (Ex i) |
|
|
C |
|
|
|
|
|
|
|
|
|
|
|
|
|
|
|
|
|
|
|
|
User configurable input/output |
|
D |
|
|
|
D |
|
|
D |
D |
D |
|
|
|
|
|
|
|
|
|
|
|
|
Pulse/frequency/switch output |
|
E |
|
|
|
E |
|
|
E |
E |
E |
|
|
|
|
|
|
|
|
|
|
|
|
Double pulse output (slave) 4) |
|
F |
|
|
|
|
|
|
F |
|
|
Pulse/frequency/switch output (Ex i) |
|
|
G |
|
|
|
|
|
|
|
|
|
|
|
|
|
|
|
|
|
|
|
|
Relay output |
|
H |
|
|
|
H |
|
|
H |
H |
H |
|
|
|
|
|
|
|
|
|
|
|
|
Current input 0/4 to 20 mA |
|
I |
|
|
|
I |
|
|
I |
I |
I |
|
|
|
|
|
|
|
|
|
|
|
|
Status input |
|
J |
|
|
|
J |
|
|
J |
J |
J |
|
|
|
|
|
|
|
|
|
|
|
|
1)A specific input or output can be assigned to a user configurable input/output → 19.
2)If double pulse output (F) is selected for output/input 2 (021), only the double pulse output (F) option is available for selection for output/input 3 (022).
3)The order code for "Output; input 4" (023) is only available for the Proline 500 – digital transmitter.
4)The double pulse output (F) option is not available for input/output 4.
Endress+Hauser |
15 |
Proline Promass F 500
Output signal |
HART current output |
|
|
|
|
|
Current output |
4 to 20 mA HART |
|
|
|
|
Current span |
Can be set to: 4 to 20 mA (active/passive) |
|
|
Ex-i, passive |
|
Open-circuit voltage |
DC 28.8 V (active) |
|
|
|
|
Maximum input voltage |
DC 30 V (passive) |
|
|
|
|
Load |
250 to 700 Ω |
|
|
|
|
Resolution |
0.38 µA |
|
|
|
|
Damping |
Configurable: 0.07 to 999 s |
|
|
|
|
Assignable measured |
• Mass flow |
|
variables |
• Volume flow |
|
|
• Corrected volume flow |
|
|
• Density |
|
|
• Reference density |
|
|
• Temperature |
|
|
• Electronic temperature |
|
|
• Oscillation frequency 0 |
|
|
• Oscillation damping 0 |
|
|
• Signal asymmetry |
|
|
• Exciter current 0 |
|
|
The range of options increases if the measuring device has one or more |
|
|
application packages. |
|
PROFIBUS PA |
|
|
|
|
|
PROFIBUS PA |
In accordance with EN 50170 Volume 2, IEC 61158-2 (MBP), galvanically |
|
|
isolated |
|
|
|
|
Data transmission |
31.25 kbit/s |
|
|
|
|
Current consumption |
10 mA |
|
|
16 mA |
|
|
|
|
Permitted supply voltage |
9 to 32 V |
|
|
|
|
Bus connection |
With integrated reverse polarity protection |
|
|
|
|
PROFIBUS DP |
|
|
|
|
|
Signal encoding |
NRZ code |
|
|
|
|
Data transfer |
9.6 kBaud…12 MBaud |
|
|
|
|
EtherNet/IP |
|
|
|
|
|
Standards |
In accordance with IEEE 802.3 |
|
|
|
|
PROFINET |
|
|
|
|
|
Standards |
In accordance with IEEE 802.3 |
|
|
|
|
FOUNDATION Fieldbus |
|
|
|
|
|
FOUNDATION Fieldbus |
H1, IEC 61158-2, galvanically isolated |
|
|
|
|
Data transfer |
31.25 kbit/s |
|
|
|
16 |
Endress+Hauser |
Proline Promass F 500
Current consumption |
10 mA |
|
|
Permitted supply voltage |
9 to 32 V |
|
|
Bus connection |
With integrated reverse polarity protection |
|
|
Modbus RS485
Physical interface |
RS485 in accordance with EIA/TIA-485 standard |
|
|
Terminating resistor |
Integrated, can be activated via DIP switches |
|
|
Current output 0/4 to 20 mA
Current output |
0/4 to 20 mA |
|
|
Maximum output values |
22.5 mA |
|
|
Current span |
Can be set to: |
|
• 4 to 20 mA (active) |
|
• 0/4 to 20 mA (passive) |
|
Ex-i, passive |
Open-circuit voltage |
DC 28.8 V (active) |
|
|
Maximum input voltage |
DC 30 V (passive) |
|
|
Load |
0 to 700 Ω |
|
|
Resolution |
0.38 µA |
|
|
Damping |
Adjustable: 0.07 to 999 s |
|
|
Assignable measured |
• Mass flow |
variables |
• Volume flow |
|
• Corrected volume flow |
|
• Density |
|
• Reference density |
|
• Temperature |
|
• Electronic temperature |
|
• Oscillation frequency 0 |
|
• Oscillation damping 0 |
|
• Signal asymmetry |
|
• Exciter current 0 |
|
The range of options increases if the measuring device has one or more |
|
application packages. |
Pulse/frequency/switch output |
|
|
|
Function |
Can be set to pulse, frequency or switch output |
|
|
Version |
Open collector |
|
Can be set to: |
|
• Active |
|
• Passive |
|
Ex-i, passive |
Maximum input values |
DC 30 V, 250 mA (passive) |
|
|
Open-circuit voltage |
DC 28.8 V (active) |
|
|
Voltage drop |
For 22.5 mA: ≤ DC 2 V |
|
|
Pulse output |
|
|
|
Maximum input values |
DC 30 V, 250 mA (passive) |
|
|
Maximum output current |
22.5 mA (active) |
|
|
Endress+Hauser |
17 |
Proline Promass F 500
Open-circuit voltage |
DC 28.8 V (active) |
|
|
Pulse width |
Adjustable: 0.05 to 2 000 ms |
|
|
Maximum pulse rate |
10 000 Impulse/s |
|
|
Pulse value |
Adjustable |
|
|
Assignable measured |
• Mass flow |
variables |
• Volume flow |
|
• Corrected volume flow |
|
|
Frequency output |
|
|
|
Maximum input values |
DC 30 V, 250 mA (passive) |
|
|
Maximum output current |
22.5 mA (active) |
|
|
Open-circuit voltage |
DC 28.8 V (active) |
|
|
Output frequency |
Adjustable: end value frequency 2 to 10 000 Hz (f max = 12 500 Hz) |
Damping |
Adjustable: 0 to 999 s |
|
|
Pulse/pause ratio |
1:1 |
|
|
Assignable measured |
• Mass flow |
variables |
• Volume flow |
|
• Corrected volume flow |
|
• Density |
|
• Reference density |
|
• Temperature |
|
• Electronic temperature |
|
• Oscillation frequency 0 |
|
• Oscillation damping 0 |
|
• Signal asymmetry |
|
• Exciter current 0 |
|
The range of options increases if the measuring device has one or more |
|
application packages. |
Switch output |
|
|
|
Maximum input values |
DC 30 V, 250 mA (passive) |
|
|
Open-circuit voltage |
DC 28.8 V (active) |
|
|
Switching behavior |
Binary, conductive or non-conductive |
|
|
Switching delay |
Adjustable: 0 to 100 s |
|
|
Number of switching |
Unlimited |
cycles |
|
|
|
Assignable functions |
• Off |
|
• On |
|
• Diagnostic behavior |
|
• Limit value |
|
– Mass flow |
|
– Volume flow |
|
– Corrected volume flow |
|
– Density |
|
– Reference density |
|
– Temperature |
|
– Totalizer 1-3 |
|
• Flow direction monitoring |
|
• Status |
|
– Partially filled pipe detection |
|
– Low flow |
|
The range of options increases if the measuring device has one or more |
|
application packages. |
18 |
Endress+Hauser |
Proline Promass F 500
Double pulse output
Function |
Double pulse |
|
|
Version |
Open collector |
|
Can be set to: |
|
• Active |
|
• Passive |
|
|
Maximum input values |
DC 30 V, 250 mA (passive) |
|
|
Open-circuit voltage |
DC 28.8 V (active) |
|
|
Voltage drop |
For 22.5 mA: ≤ DC 2 V |
|
|
Output frequency |
Adjustable: 0 to 1 000 Hz |
|
|
Damping |
Adjustable: 0 to 999 s |
|
|
Pulse/pause ratio |
1:1 |
|
|
Assignable measured |
• Mass flow |
variables |
• Volume flow |
|
• Corrected volume flow |
|
• Density |
|
• Reference density |
|
• Temperature |
|
The range of options increases if the measuring device has one or more |
|
application packages. |
Relay output |
|
|
|
Function |
Switch output |
|
|
Version |
Relay output, galvanically isolated |
|
|
Switching behavior |
Can be set to: |
|
• NO (normally open), factory setting |
|
• NC (normally closed) |
|
|
Maximum switching |
• DC 30 V, 0.1 A |
capacity (passive) |
• AC 30 V, 0.5 A |
|
|
Assignable functions |
• Off |
|
• On |
|
• Diagnostic behavior |
|
• Limit value |
|
– Mass flow |
|
– Volume flow |
|
– Corrected volume flow |
|
– Density |
|
– Reference density |
|
– Temperature |
|
– Totalizer 1-3 |
|
• Flow direction monitoring |
|
• Status |
|
– Partially filled pipe detection |
|
– Low flow |
|
The range of options increases if the measuring device has one or more |
|
application packages. |
User configurable input/output
One specific input or output is assigned to a user-configurable input/output (configurable I/O) during device commissioning.
The following inputs and outputs are available for assignment:
•Choice of current output: 4 to 20 mA (active), 0/4 to 20 mA (passive)
•Pulse/frequency/switch output
•Choice of current input: 4 to 20 mA (active), 0/4 to 20 mA (passive)
•Status input
Endress+Hauser |
19 |
|
|
Proline Promass F 500 |
|
|
|
|
The technical values correspond to those of the inputs and outputs described in this section. |
|
|
|
|
Signal on alarm |
Depending on the interface, failure information is displayed as follows: |
|
|
HART current output |
|
|
|
|
|
Device diagnostics |
Device condition can be read out via HART Command 48 |
|
|
|
|
PROFIBUS PA |
|
|
|
|
|
Status and alarm |
Diagnostics in accordance with PROFIBUS PA Profile 3.02 |
|
messages |
|
|
|
|
|
Error current FDE (Fault |
0 mA |
|
Disconnection Electronic) |
|
|
|
|
|
PROFIBUS DP |
|
|
|
|
|
Status and alarm |
Diagnostics in accordance with PROFIBUS PA Profile 3.02 |
|
messages |
|
|
|
|
|
EtherNet/IP |
|
|
|
|
|
Device diagnostics |
Device condition can be read out in Input Assembly |
|
|
|
|
PROFINET |
|
|
|
|
|
Device diagnostics |
According to "Application Layer protocol for decentralized periphery", Version 2.3 |
|
|
|
|
FOUNDATION Fieldbus |
|
|
|
|
|
Status and alarm |
Diagnostics in accordance with FF-891 |
|
messages |
|
|
|
|
|
Error current FDE (Fault |
0 mA |
|
Disconnection Electronic) |
|
|
|
|
|
Modbus RS485 |
|
|
|
|
|
Failure mode |
Choose from: |
|
|
• NaN value instead of current value |
|
|
• Last valid value |
|
|
|
Current output 0/4 to 20 mA
4 to 20 mA
Failure mode |
Choose from: |
|
• 4 to 20 mA in accordance with NAMUR recommendation NE 43 |
|
• 4 to 20 mA in accordance with US |
|
• Min. value: 3.59 mA |
|
• Max. value: 22.5 mA |
|
• Freely definable value between: 3.59 to 22.5 mA |
|
• Actual value |
|
• Last valid value |
|
|
20 |
Endress+Hauser |
Proline Promass F 500
0 to 20 mA
Failure mode |
Choose from: |
|
• Maximum alarm: 22 mA |
|
• Freely definable value between: 0 to 20.5 mA |
|
|
Pulse/frequency/switch output
Pulse output
Failure mode |
Choose from: |
|
• Actual value |
|
• No pulses |
|
|
Frequency output |
|
|
|
Failure mode |
Choose from: |
|
• Actual value |
|
• 0 Hz |
|
• Defined value (f max 2 to 12 500 Hz) |
Switch output |
|
|
|
Failure mode |
Choose from: |
|
• Current status |
|
• Open |
|
• Closed |
|
|
Relay output
Failure mode |
Choose from: |
|
• Current status |
|
• Open |
|
• Closed |
|
|
Local display
Plain text display |
With information on cause and remedial measures |
|
|
Backlight |
Red backlighting indicates a device error. |
|
|
Status signal as per NAMUR recommendation NE 107
Interface/protocol
•Via digital communication:
–HART protocol
–FOUNDATION Fieldbus
–PROFIBUS PA
–PROFIBUS DP
–Modbus RS485
–EtherNet/IP
–PROFINET
•Via service interface
–CDI-RJ45 service interface
–WLAN interface
Plain text display |
With information on cause and remedial measures |
|
|
Additional information on remote operation → 113
Endress+Hauser |
21 |
Proline Promass F 500
Web server
Plain text display |
With information on cause and remedial measures |
|
|
Light emitting diodes (LED)
Status information |
Status indicated by various light emitting diodes |
The following information is displayed depending on the device version:
•Supply voltage active
•Data transmission active
•Device alarm/error has occurred
•EtherNet/IP network available
•EtherNet/IP connection established
•PROFINET network available
•PROFINET connection established
•PROFINET blinking feature
Ex connection data |
Safety-related values |
|
|
|
|
|
|
|
|
|
|
|
|
|
|
|
|
|
|
|
|
|
|
|
|
|
|
|
Order code for |
|
Output type |
|
|
|
|
Safety-related values |
|||||
|
"Output; input 1" |
|
|
|
|
|
|
"Output; input 1" |
|
|
|||
|
|
|
|
|
26 (+) |
|
|
|
27 (–) |
||||
|
|
|
|
|
|
|
|
||||||
|
|
|
|
|
|
|
|
|
|
|
|
|
|
|
Option BA |
|
Current output |
|
UN = 30 VDC |
|
|
|
|
|
|
|
|
|
|
|
4 to 20 mA HART |
|
UM = 250 VAC |
|
|
|
|
|
|
|
|
|
Option GA |
|
PROFIBUS PA |
|
UN = 30 VDC |
|
|
|
|
|
|
|
|
|
|
|
|
|
UM = 250 VAC |
|
|
|
|
|
|
|
|
|
Option LA |
|
PROFIBUS DP |
|
UN = 30 VDC |
|
|
|
|
|
|
|
|
|
|
|
|
|
UM = 250 VAC |
|
|
|
|
|
|
|
|
|
Option MA |
|
Modbus RS485 |
|
UN = 30 VDC |
|
|
|
|
|
|
|
|
|
|
|
|
|
UM = 250 VAC |
|
|
|
|
|
|
|
|
|
Option SA |
|
FOUNDATION Fieldbus |
|
UN = 30 VDC |
|
|
|
|
|
|
|
|
|
|
|
|
|
UM = 250 VAC |
|
|
|
|
|
|
|
|
|
Option NA |
|
EtherNet/IP |
|
UN = 30 VDC |
|
|
|
|
|
|
|
|
|
|
|
|
|
UM = 250 VAC |
|
|
|
|
|
|
|
|
|
Option RA |
|
PROFINET |
|
UN = 30 VDC |
|
|
|
|
|
|
|
|
|
|
|
|
|
UM = 250 VAC |
|
|
|
|
|
|
|
|
|
|
|
|
|
|
|
|
|
|
|
|
||
|
Order code for |
Output type |
|
|
|
Safety-related values |
|
|
|||||
|
"Output; input 2"; |
|
|
Output; input 2 |
|
Output; input 3 |
|
Output; input |
|||||
|
|
|
|
|
|||||||||
|
"Output; input 3" |
|
|
|
|
||||||||
|
|
|
|
|
|
|
|
|
|
|
4 1) |
||
|
"Output; input 4" |
|
|
|
|
|
|
|
|
|
|
||
|
|
|
|
24 (+) |
25 (–) |
|
22 (+) |
23 (–) |
|
20 (+) |
21 (–) |
||
|
|
|
|
|
|
|
|
|
|
|
|
||
|
Option B |
Current output |
UN = 30 VDC |
|
|
|
|
|
|
|
|||
|
|
4 to 20 mA |
UM = 250 VAC |
|
|
|
|
|
|
|
|||
|
Option D |
User configurable input/ |
UN = 30 VDC |
|
|
|
|
|
|
|
|||
|
|
output |
UM = 250 VAC |
|
|
|
|
|
|
|
|||
|
Option E |
Pulse/frequency/switch |
UN = 30 VDC |
|
|
|
|
|
|
|
|||
|
|
output |
UM = 250 VAC |
|
|
|
|
|
|
|
|||
|
Option F |
Double pulse output |
UN = 30 VDC |
|
|
|
|
|
|
|
|||
|
|
|
|
UM = 250 VAC |
|
|
|
|
|
|
|
||
|
Option H |
Relay output |
UN = 30 VDC |
|
|
|
|
|
|
|
|||
|
|
|
|
IN =100 mADC/500 mAAC |
|
|
|||||||
|
|
|
|
UM = 250 VAC |
|
|
|
|
|
|
|
22 |
Endress+Hauser |
Proline Promass F 500
Order code for |
Output type |
|
|
Safety-related values |
|
|
|||
"Output; input 2"; |
|
Output; input 2 |
|
Output; input 3 |
|
Output; input |
|||
|
|
|
|||||||
"Output; input 3" |
|
|
|
||||||
|
|
|
|
|
|
|
4 1) |
||
"Output; input 4" |
|
|
|
|
|
|
|
||
|
|
24 (+) |
25 (–) |
|
22 (+) |
23 (–) |
|
20 (+) |
21 (–) |
|
|
|
|
|
|
|
|
|
|
Option I |
Current input |
UN = 30 VDC |
|
|
|
|
|
|
|
|
4 to 20 mA |
UM = 250 VAC |
|
|
|
|
|
|
|
Option J |
Status input |
UN = 30 VDC |
|
|
|
|
|
|
|
|
|
UM = 250 VAC |
|
|
|
|
|
|
1)The order code "Output; input 4" is only available for the Proline 500 – digital transmitter.
Intrinsically safe values
Order code for |
Output type |
Intrinsically safe values |
||
"Output; input 1" |
|
|
"Output; input 1" |
|
|
|
26 (+) |
|
27 (–) |
|
|
|
||
|
|
|
|
|
Option CA |
Current output |
Ui = 30 V |
|
|
|
4 to 20 mA HART Ex i |
li = 100 mA |
|
|
|
|
Pi = 1.25 W |
|
|
|
|
Li = 0 |
|
|
|
|
Ci = 0 |
|
|
Option HA |
PROFIBUS PA Ex i |
Ex ia 1) |
|
Ex ic 2) |
|
|
Ui = 30 V |
|
Ui = 32 V |
|
|
li = 570 mA |
|
li = 570 mA |
|
|
Pi = 8.5 W |
|
Pi = 8.5 W |
|
|
Li = 10 µH |
|
Li = 10 µH |
|
|
Ci = 5 nF |
|
Ci = 5 nF |
Option TA |
FOUNDATION Fieldbus |
Ex ia 1) |
|
Ex ic 2) |
|
Ex i |
Ui = 30 V |
|
Ui = 32 V |
|
|
li = 570 mA |
|
li = 570 mA |
|
|
Pi = 8.5 W |
|
Pi = 8.5 W |
|
|
Li = 10 µH |
|
Li = 10 µH |
|
|
Ci = 5 nF |
|
Ci = 5 nF |
1)Only available for the Zone 1; Class I, Division 1 version
2)Only available for the Zone 2; Class I, Division 2 version and only for the Proline 500 – digital transmitter
Order code for |
Output type |
Intrinsically safe values or NIFW values |
|||||||
"Output; input 2"; |
|
Output; input 2 |
|
Output; input 3 |
|
Output; input 4 1) |
|||
|
|
|
|||||||
"Output; input 3" |
|
|
|
||||||
"Output; input 4" |
|
24 (+) |
25 (–) |
|
22 (+) |
23 (–) |
|
20 (+) |
21 (–) |
|
|
|
|
|
|
|
|
|
|
Option C |
Current output |
Ui = 30 V |
|
|
|
|
|
|
|
|
4 to 20 mA Ex i |
li = 100 mA |
|
|
|
|
|
|
|
|
|
Pi = 1.25 W |
|
|
|
|
|
|
|
|
|
Li = 0 |
|
|
|
|
|
|
|
|
|
Ci = 0 |
|
|
|
|
|
|
|
Option G |
Pulse/frequency/switch |
Ui = 30 V |
|
|
|
|
|
|
|
|
output Ex i |
li = 100 mA |
|
|
|
|
|
|
|
|
|
Pi = 1.25 W |
|
|
|
|
|
|
|
|
|
Li = 0 |
|
|
|
|
|
|
|
|
|
Ci = 0 |
|
|
|
|
|
|
1)The order code "Output; input 4" is only available for the Proline 500 – digital transmitter.
Low flow cut off |
The switch points for low flow cut off are user-selectable. |
|
|
Galvanic isolation |
The outputs are galvanically isolated from one another and from earth (PE). |
Endress+Hauser |
23 |
Proline Promass F 500
Protocol-specific data |
HART |
|
|
|
|
|
Manufacturer ID |
0x11 |
|
|
|
|
Device type ID |
0x3B |
|
|
|
|
HART protocol revision |
7 |
|
|
|
|
Device description files |
Information and files under: |
|
(DTM, DD) |
www.endress.com |
|
|
|
|
HART load |
Min. 250 Ω |
|
|
|
|
System integration |
Information on system integration: Operating Instructions → 135. |
|
|
• Measured variables via HART protocol |
|
|
• Burst Mode functionality |
|
|
|
PROFIBUS PA
Manufacturer ID |
0x11 |
|
|
Ident number |
0x156D |
|
|
Profile version |
3.02 |
|
|
Device description files (GSD, |
Information and files under: |
DTM, DD) |
• www.endress.com |
|
• www.profibus.org |
|
|
Supported functions |
• Identification & Maintenance |
|
Simplest device identification on the part of the control system and |
|
nameplate |
|
• PROFIBUS upload/download |
|
Reading and writing parameters is up to ten times faster with PROFIBUS |
|
upload/download |
|
• Condensed status |
|
Simplest and self-explanatory diagnostic information by categorizing |
|
diagnostic messages that occur |
|
|
Configuration of the device |
• DIP switches on the I/O electronics module |
address |
• Local display |
|
• Via operating tools (e.g. FieldCare) |
|
|
Compatibility with |
If the device is replaced, the measuring device Promass 500 supports the |
earlier model |
compatibility of the cyclic data with previous models. It is not necessary to |
|
adjust the engineering parameters of the PROFIBUS network with the |
|
Promass 500 GSD file. |
|
Earlier models: |
|
• Promass 80 PROFIBUS PA |
|
– ID No.: 1528 (hex) |
|
– Extended GSD file: EH3x1528.gsd |
|
– Standard GSD file: EH3_1528.gsd |
|
• Promass 83 PROFIBUS PA |
|
– ID No.: 152A (hex) |
|
– Extended GSD file: EH3x152A.gsd |
|
– Standard GSD file: EH3_152A.gsd |
|
Description of the function scope of compatibility: |
|
Operating Instructions → 135. |
System integration |
Information regarding system integration: Operating Instructions → 135. |
|
• Cyclic data transmission |
|
• Block model |
|
• Description of the modules |
|
|
PROFIBUS DP |
|
|
|
Manufacturer ID |
0x11 |
|
|
Ident number |
0x156F |
|
|
24 |
Endress+Hauser |
Proline Promass F 500
Profile version |
3.02 |
|
|
Device description files (GSD, |
Information and files under: |
DTM, DD) |
• www.endress.com |
|
On the product page for the device: Documents/Software → Device drivers |
|
• www.profibus.org |
|
|
Supported functions |
• Identification & Maintenance |
|
Simplest device identification on the part of the control system and |
|
nameplate |
|
• PROFIBUS upload/download |
|
Reading and writing parameters is up to ten times faster with PROFIBUS |
|
upload/download |
|
• Condensed status |
|
Simplest and self-explanatory diagnostic information by categorizing |
|
diagnostic messages that occur |
|
|
Configuration of the device |
• DIP switches on the I/O electronics module |
address |
• Via operating tools (e.g. FieldCare) |
|
|
System integration |
Information regarding system integration: Operating Instructions → 135. |
|
• Cyclic data transmission |
|
• Block model |
|
• Description of the modules |
|
|
EtherNet/IP
Protocol |
• The CIP Networks Library Volume 1: Common Industrial Protocol |
|
• The CIP Networks Library Volume 2: EtherNet/IP Adaptation of CIP |
|
|
Communication type |
• 10Base-T |
|
• 100Base-TX |
|
|
Device profile |
Generic device (product type: 0x2B) |
|
|
Manufacturer ID |
0x11 |
|
|
Device type ID |
0x103B |
|
|
Baud rates |
Automatic ¹ ⁄ Mbit with half-duplex and full-duplex detection |
|
|
Polarity |
Auto-polarity for automatic correction of crossed TxD and RxD pairs |
|
|
Supported CIP connections |
Max. 3 connections |
|
|
Explicit connections |
Max. 6 connections |
|
|
I/O connections |
Max. 6 connections (scanner) |
|
|
Configuration options for |
• DIP switches on the electronics module for IP addressing |
measuring device |
• Manufacturer-specific software (FieldCare) |
|
• Add-on Profile Level 3 for Rockwell Automation control systems |
|
• Web browser |
|
• Electronic Data Sheet (EDS) integrated in the measuring device |
|
|
Configuration of the EtherNet |
• Speed: 10 MBit, 100 MBit, auto (factory setting) |
interface |
• Duplex: half-duplex, full-duplex, auto (factory setting) |
|
|
Configuration of the device |
• DIP switches on the electronics module for IP addressing (last octet) |
address |
• DHCP |
|
• Manufacturer-specific software (FieldCare) |
|
• Add-on Profile Level 3 for Rockwell Automation control systems |
|
• Web browser |
|
• EtherNet/IP tools, e.g. RSLinx (Rockwell Automation) |
|
|
Device Level Ring (DLR) |
Yes |
|
|
System integration |
Information regarding system integration: Operating Instructions |
|
→ 135. |
|
• Cyclic data transmission |
|
• Block model |
|
• Input and output groups |
|
|
Endress+Hauser |
25 |
|
|
Proline Promass F 500 |
|
|
|
|
PROFINET |
|
|
|
|
|
Protocol |
"Application layer protocol for decentral device periphery and distributed |
|
|
automation", version 2.3 |
|
|
|
|
Communication type |
100 MBit/s |
|
|
|
|
Conformity class |
Conformance Class B |
|
|
|
|
Netload Class |
Netload Class II |
|
|
|
|
Baud rates |
Automatic 100 Mbit/s with full-duplex detection |
|
|
|
|
Cycle times |
From 8 ms |
|
|
|
|
Polarity |
Auto-polarity for automatic correction of crossed TxD and RxD pairs |
|
|
|
|
Media Redundancy Protocol |
Yes |
|
(MRP) |
|
|
|
|
|
Device profile |
Application interface identifier 0xF600 |
|
|
Generic device |
|
|
|
|
Manufacturer ID |
0x11 |
|
|
|
|
Device type ID |
0x843B |
|
|
|
|
Device description files (GSD, |
Information and files under: |
|
DTM, DD) |
• www.endress.com |
|
|
On the product page for the device: Documents/Software → Device drivers |
|
|
• www.profibus.org |
|
|
|
|
Supported connections |
• 1 x AR (IO Controller AR) |
|
|
• 1 x AR (IO-Supervisor Device AR connection allowed) |
|
|
• 1 x Input CR (Communication Relation) |
|
|
• 1 x Output CR (Communication Relation) |
|
|
• 1 x Alarm CR (Communication Relation) |
|
|
|
|
Configuration options for |
• DIP switches on the electronics module, for device name assignment (last |
|
measuring device |
part) |
|
|
• Manufacturer-specific software (FieldCare, DeviceCare) |
|
|
• Web browser |
|
|
• Device master file (GSD), can be read out via the integrated Web server of |
|
|
the measuring device |
|
|
|
|
Configuration of the |
• DIP switches on the electronics module, for device name assignment (last |
|
device name |
part) |
|
|
• DCP protocol |
|
|
• Process Device Manager (PDM) |
|
|
• Integrated Web server |
|
|
|
|
Supported functions |
• Identification & Maintenance |
|
|
Simple device identification via: |
|
|
– Control system |
|
|
– Nameplate |
|
|
• Measured value status |
|
|
The process variables are communicated with a measured value status |
|
|
• Blinking feature via the onsite display for simple device identification and |
|
|
assignment |
|
|
• Device operation via operating tools (e.g. FieldCare, DeviceCare, SIMATIC |
|
|
PDM) |
|
|
|
|
System integration |
Information regarding system integration: Operating Instructions → 135. |
|
|
• Cyclic data transmission |
|
|
• Overview and description of the modules |
|
|
• Status coding |
|
|
• Startup configuration |
|
|
• Factory setting: |
|
|
|
|
FOUNDATION Fieldbus |
|
|
|
|
|
Manufacturer ID |
0x452B48 (hex) |
|
|
|
|
Ident number |
0x103B (hex) |
|
|
|
26 |
Endress+Hauser |
Proline Promass F 500
Device revision |
1 |
|
|
|
|
DD revision |
Information and files under: |
|
|
• www.endress.com |
|
CFF revision |
||
• www.fieldbus.org |
||
|
||
|
|
|
Interoperability Test Kit (ITK) |
Version 6.2.0 |
|
|
|
|
ITK Test Campaign Number |
Information: |
|
|
• www.endress.com |
|
|
• www.fieldbus.org |
|
|
|
|
Link Master capability (LAS) |
Yes |
|
|
|
|
Choice of "Link Master" and |
Yes |
|
"Basic Device" |
Factory setting: Basic Device |
|
|
|
|
Node address |
Factory setting: 247 (0xF7) |
|
|
|
|
Supported functions |
The following methods are supported: |
|
|
• Restart |
|
|
• ENP Restart |
|
|
• Diagnostic |
|
|
• Set to OOS |
|
|
• Set to AUTO |
|
|
• Read trend data |
|
|
• Read event logbook |
|
|
|
|
Virtual Communication Relationships (VCRs) |
||
|
|
|
Number of VCRs |
44 |
|
|
|
|
Number of link objects in VFD |
50 |
|
|
|
|
Permanent entries |
1 |
|
|
|
|
Client VCRs |
0 |
|
|
|
|
Server VCRs |
10 |
|
|
|
|
Source VCRs |
43 |
|
|
|
|
Sink VCRs |
0 |
|
|
|
|
Subscriber VCRs |
43 |
|
|
|
|
Publisher VCRs |
43 |
|
|
|
|
Device Link Capabilities |
|
|
|
|
|
Slot time |
4 |
|
|
|
|
Min. delay between PDU |
8 |
|
|
|
|
Max. response delay |
16 |
|
|
|
|
System integration |
Information regarding system integration: Operating Instructions → 135. |
|
|
• Cyclic data transmission |
|
|
• Description of the modules |
|
|
• Execution times |
|
|
• Methods |
|
|
|
Modbus RS485
Protocol |
Modbus Applications Protocol Specification V1.1 |
|
|
Response times |
• Direct data access: typically 25 to 50 ms |
|
• Auto-scan buffer (data range): typically 3 to 5 ms |
|
|
Device type |
Slave |
|
|
Slave address range |
1 to 247 |
|
|
Broadcast address range |
0 |
|
|
Endress+Hauser |
27 |
Proline Promass F 500
|
Function codes |
|
|
• 03: Read holding register |
|
|
|
|
|
|
|||||
|
|
|
|
|
|
• 04: Read input register |
|
|
|
|
|
|
|||
|
|
|
|
|
|
• 06: Write single registers |
|
|
|
|
|
|
|||
|
|
|
|
|
|
• 08: Diagnostics |
|
|
|
|
|
|
|
||
|
|
|
|
|
|
• 16: Write multiple registers |
|
|
|
|
|
|
|||
|
|
|
|
|
|
• 23: Read/write multiple registers |
|
|
|
|
|
||||
|
|
|
|
|
|
|
|
|
|
|
|
|
|
||
|
Broadcast messages |
Supported by the following function codes: |
|
|
|
|
|
||||||||
|
|
|
|
|
|
• 06: Write single registers |
|
|
|
|
|
|
|||
|
|
|
|
|
|
• 16: Write multiple registers |
|
|
|
|
|
|
|||
|
|
|
|
|
|
• 23: Read/write multiple registers |
|
|
|
|
|
||||
|
|
|
|
|
|
|
|
|
|
|
|
|
|
|
|
|
Supported baud rate |
• 1 200 BAUD |
|
|
|
|
|
|
|
||||||
|
|
|
|
|
|
• 2 400 BAUD |
|
|
|
|
|
|
|
||
|
|
|
|
|
|
• 4 800 BAUD |
|
|
|
|
|
|
|
||
|
|
|
|
|
|
• 9 600 BAUD |
|
|
|
|
|
|
|
||
|
|
|
|
|
|
• 19 200 BAUD |
|
|
|
|
|
|
|
||
|
|
|
|
|
|
• 38 400 BAUD |
|
|
|
|
|
|
|
||
|
|
|
|
|
|
• 57 600 BAUD |
|
|
|
|
|
|
|
||
|
|
|
|
|
|
• 115 200 BAUD |
|
|
|
|
|
|
|
||
|
|
|
|
|
|
|
|
|
|
|
|
|
|
|
|
|
Data transfer mode |
|
|
• ASCII |
|
|
|
|
|
|
|
|
|
||
|
|
|
|
|
|
• RTU |
|
|
|
|
|
|
|
|
|
|
|
|
|
|
|
|
|
|
|
|
|
|
|
||
|
Data access |
|
|
Each device parameter can be accessed via Modbus RS485. |
|
||||||||||
|
|
|
|
|
|
For Modbus register information |
|
|
|
|
|
||||
|
Compatibility with |
|
|
If the device is replaced, the measuring device Promass 500 supports the |
|||||||||||
|
earlier model |
|
|
compatibility of the Modbus registers for the process variables and the |
|||||||||||
|
|
|
|
|
|
diagnostic information with the previous model Promass 83. It is not |
|
||||||||
|
|
|
|
|
|
necessary to change the engineering parameters in the automation system. |
|||||||||
|
|
|
|
|
|
Description of the function scope of compatibility: |
|
|
|
||||||
|
|
|
|
|
|
Operating Instructions → 135. |
|
|
|
|
|
||||
|
System integration |
|
|
Information on system integration: Operating Instructions → 135. |
|
||||||||||
|
|
|
|
|
|
• Modbus RS485 information |
|
|
|
|
|
|
|||
|
|
|
|
|
|
• Function codes |
|
|
|
|
|
|
|
||
|
|
|
|
|
|
• Register information |
|
|
|
|
|
|
|||
|
|
|
|
|
|
• Response time |
|
|
|
|
|
|
|
||
|
|
|
|
|
|
• Modbus data map |
|
|
|
|
|
|
|
||
|
|
|
|
|
|
|
|
|
|
|
|
|
|
|
|
|
Power supply |
|
|
|
|
|
|
|
|
|
|
||||
|
|
|
|
|
|
|
|
|
|||||||
Terminal assignment |
Transmitter: supply voltage, input/outputs |
|
|
|
|
|
|
|
|||||||
|
HART |
|
|
|
|
|
|
|
|
|
|
|
|
|
|
|
|
|
|
|
|
|
|
|
|
|
|
||||
|
Supply voltage |
|
Input/output |
Input/output |
Input/output |
Input/output |
|||||||||
|
|
|
|
1 |
|
2 |
|
|
3 |
4 |
|
||||
|
|
|
|
|
|
|
|
|
|
|
|
|
|
|
|
|
1 (+) |
2 (–) |
|
26 (+) |
|
27 (–) |
24 (+) |
|
25 (–) |
22 (+) |
|
23 (–) |
20 (+) |
|
21 (–) |
|
|
|
|
|
|
|
|
|
|
|
|
||||
|
|
|
|
The terminal assignment depends on the specific device version ordered → 15. |
|||||||||||
|
|
|
|
|
|
|
|
|
|
|
|
|
|
|
|
|
FOUNDATION Fieldbus |
|
|
|
|
|
|
|
|
|
|
||||
|
|
|
|
|
|
||||||||||
|
Supply voltage |
|
Input/output |
Input/output |
Input/output |
Input/output |
|||||||||
|
|
|
|
1 |
|
2 |
|
|
3 |
4 |
|
||||
|
|
|
|
|
|
|
|
|
|
|
|
|
|
|
|
|
1 (+) |
2 (–) |
|
26 (A) |
|
27 (B) |
24 (+) |
|
25 (–) |
22 (+) |
|
23 (–) |
20 (+) |
|
21 (–) |
|
|
|
|
|
|
|
|
|
|
|
|
||||
|
|
|
|
The terminal assignment depends on the specific device version ordered → 15. |
|||||||||||
|
|
|
|
|
|
|
|
|
|
|
|
|
|
|
|
28 |
Endress+Hauser |
Proline Promass F 500
PROFIBUS PA
Supply voltage |
Input/output |
Input/output |
Input/output |
Input/output |
|||||||||
|
|
|
1 |
|
2 |
|
3 |
|
4 |
||||
|
|
|
|
|
|
|
|
|
|
|
|
|
|
1 (+) |
2 (–) |
26 (B) |
|
27 (A) |
24 (+) |
|
25 (–) |
22 (+) |
|
23 (–) |
20 (+) |
|
21 (–) |
|
|
|
|
|
|
|
|
|
|
||||
|
|
The terminal assignment depends on the specific device version ordered → 15. |
|||||||||||
|
|
|
|
|
|
|
|
|
|
|
|
|
|
PROFIBUS DP
Supply voltage |
Input/output |
Input/output |
Input/output |
Input/output |
|||||||||
|
|
|
1 |
|
2 |
|
3 |
|
4 |
||||
|
|
|
|
|
|
|
|
|
|
|
|
|
|
1 (+) |
2 (–) |
26 (B) |
|
27 (A) |
24 (+) |
|
25 (–) |
22 (+) |
|
23 (–) |
20 (+) |
|
21 (–) |
|
|
|
|
|
|
|
|
|
|
||||
|
|
The terminal assignment depends on the specific device version ordered → 15. |
|||||||||||
|
|
|
|
|
|
|
|
|
|
|
|
|
|
Modbus RS485
Supply voltage |
Input/output |
Input/output |
Input/output |
Input/output |
|||||||||
|
|
|
1 |
|
2 |
|
3 |
|
4 |
||||
|
|
|
|
|
|
|
|
|
|
|
|
|
|
1 (+) |
2 (–) |
26 (B) |
|
27 (A) |
24 (+) |
|
25 (–) |
22 (+) |
|
23 (–) |
20 (+) |
|
21 (–) |
|
|
|
|
|
|
|
|
|
|
||||
|
|
The terminal assignment depends on the specific device version ordered → 15. |
|||||||||||
|
|
|
|
|
|
|
|
|
|
|
|
|
|
PROFINET
Supply voltage |
Input/output |
Input/output |
Input/output |
Input/output |
|||||||
|
|
1 |
|
2 |
|
3 |
|
4 |
|||
|
|
|
|
|
|
|
|
|
|
|
|
1 (+) |
2 (–) |
PROFINET |
24 (+) |
|
25 (–) |
22 (+) |
|
23 (–) |
20 (+) |
|
21 (–) |
|
|
(RJ45 connector) |
|
|
|
|
|
|
|
|
|
|
|
The terminal assignment depends on the specific device version |
|||||||||
|
|
|
|||||||||
|
|
|
|
|
|
ordered → 15. |
|
|
|
||
|
|
|
|
|
|
|
|
|
|
|
|
EtherNet/IP
Supply voltage |
Input/output |
Input/output |
Input/output |
Input/output |
|||||||
|
|
1 |
|
2 |
|
3 |
|
4 |
|||
|
|
|
|
|
|
|
|
|
|
|
|
1 (+) |
2 (–) |
EtherNet/IP |
24 (+) |
|
25 (–) |
22 (+) |
|
23 (–) |
20 (+) |
|
21 (–) |
|
|
(RJ45 connector) |
|
|
|
|
|
|
|
|
|
|
|
The terminal assignment depends on the specific device version |
|||||||||
|
|
|
|||||||||
|
|
|
|
|
|
ordered → 15. |
|
|
|
||
|
|
|
|
|
|
|
|
|
|
|
|
Transmitter and sensor connection housing: connecting cable
The sensor and transmitter, which are mounted in separate locations, are interconnected by a connecting cable. The cable is connected via the sensor connection housing and the transmitter housing.
Terminal assignment and connection of the connecting cable:
•Proline 500 – digital→ 33
•Proline 500 → 34
Device plugs available |
Device plugs may not be used in hazardous areas! |
Device plugs for fieldbus systems:
Order code for "Input; output 1"
•Option SA "FOUNDATION Fieldbus" → 30
•Option GA "PROFIBUS PA" → 30
•Option RA "PROFINET" → 30
•Option NA "EtherNet/IP" → 30
Device plug for connecting to the service interface:
Order code for "Accessory mounted"
option NB, adapter RJ45 M12 (service interface) → 31
Endress+Hauser |
29 |
Proline Promass F 500
Order code for "Input; output 1", option SA "FOUNDATION Fieldbus"
Order code for |
Cable entry/connection → 34 |
|
|
"Electrical connection" |
2 |
|
3 |
|
|||
|
|
|
|
M, 3, 4, 5 |
7/8" connector |
|
– |
|
|
|
|
Order code for "Input; output 1", option GA "PROFIBUS PA"
Order code for |
Cable entry/connection → 34 |
|
|
"Electrical connection" |
2 |
|
3 |
|
|||
|
|
|
|
L, N, P, U |
Connector M12 × 1 |
|
– |
|
|
|
|
Order code for "Input; output 1", option RA "PROFINET"
Order code for |
Cable entry/connection → 34 |
|
"Electrical connection" |
2 |
|
3 |
||
|
|
|
L, N, P, U |
Connector M12 × 1 |
– |
|
|
|
R 1) 2), S 1) 2), T 1) 2), V 1) 2) |
Connector M12 × 1 |
Connector M12 × 1 |
1)Cannot be combined with an external WLAN antenna (order code for "Enclosed accessories", option P8) of an RJ45 M12 adapter for the service interface (order code for "Accessories mounted", option NB) or of the remote display and operating module DKX001.
2)Suitable for integrating the device in a ring topology.
Order code for "Input; output 1", option NA "EtherNet/IP"
Order code for |
Cable entry/connection → 34 |
|
"Electrical connection" |
2 |
|
3 |
||
|
|
|
L, N, P, U |
Connector M12 × 1 |
– |
|
|
|
R 1) 2), S 1) 2), T 1) 2), V 1) 2) |
Connector M12 × 1 |
Connector M12 × 1 |
1)Cannot be combined with an external WLAN antenna (order code for "Enclosed accessories", option P8) of an RJ45 M12 adapter for the service interface (order code for "Accessories mounted", option NB) or of the remote display and operating module DKX001
2)Suitable for integrating the device in a ring topology.
Order code for "Accessory mounted", option NB "Adapter RJ45 M12 (service interface)"
Order code |
Cable entry/coupling → 34 |
|
|
"Accessory mounted" |
Cable entry |
|
Cable entry |
|
|||
|
2 |
|
3 |
|
|
|
|
NB |
Plug M12 × 1 |
|
– |
|
|
|
|
Pin assignment, device plug |
FOUNDATION Fieldbus |
|
|
|
|
|
|||||
|
|
|
|
|
|
|
|
|
|
|
|
|
|
|
|
|
|
|
Pin |
|
Assignment |
Coding |
Plug/socket |
|
2 |
|
|
|
3 |
|
|
|
|
|
|
|
|
|
|
|
|
|
1 |
+ |
Signal + |
A |
Plug |
|
|
|
|
|
|
|
|
|
|
|
|
|
1 |
|
|
|
4 |
|
2 |
- |
Signal – |
|
|
|
|
|
|
|
|
||||||
|
|
|
|
|
|
|
|
|
|
|
|
|
|
|
|
|
|
|
3 |
|
Grounding |
|
|
|
|
|
|
|
|
|
|
|
|
|
|
|
|
|
|
|
|
|
4 |
|
Not assigned |
|
|
|
|
|
|
|
|
|
|
|
|
|
|
30 |
Endress+Hauser |