Endress+Hauser F 500 Specifications

4 (2)

TI01222D/06/EN/04.18

71402584

2018-06-01

Products

Solutions

Services

 

 

 

Technical Information

Proline Promass F 500

Coriolis flowmeter

The flowmeter offering premium accuracy and robustness, available as remote version with up to 4 I/Os

Application

Measuring principle operates independently of physical fluid properties such as viscosity or density

Highest measurement performance for liquids and gases under varying, demanding process conditions

Device properties

Mass flow: measured error ±0.05 % (PremiumCal)

Medium temperature: –196 to +350 °C (–320 to +662 °F)

Nominal diameter: DN 8 to 250 (³⁄ to 10")

Remote version with up to 4 I/Os

Backlit display with touch control and WLAN access

Standard cable between sensor and transmitter

Your benefits

Highest process safety – immune to fluctuating and harsh environments

Fewer process measuring points – multivariable measurement (flow, density, temperature)

Space saving installation – no inlet/outlet run needs

Full access to process and diagnostic information – numerous, freely combinable I/Os and fieldbuses

Reduced complexity and variety – freely configurable I/O functionality

Integrated verification – Heartbeat Technology

Proline Promass F 500

Table of contents

About this document . . . . . . . . . . . . . . . . . . . . . . . . 4

Symbols used . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . 4

Function and system design . . . . . . . . . . . . . . . . . . . 5

Measuring principle . . . . . . . . . . . . . . . . . . . . . . . . . . . . 5 Measuring system . . . . . . . . . . . . . . . . . . . . . . . . . . . . . 7 Equipment architecture . . . . . . . . . . . . . . . . . . . . . . . . . 9 Safety . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . 9

Input . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . 12

Measured variable . . . . . . . . . . . . . . . . . . . . . . . . . . . . 12 Measuring range . . . . . . . . . . . . . . . . . . . . . . . . . . . . . 12 Operable flow range . . . . . . . . . . . . . . . . . . . . . . . . . . . 13 Input signal . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . 13

Output . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . 15

Output and input variants . . . . . . . . . . . . . . . . . . . . . . . 15 Output signal . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . 16 Signal on alarm . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . 20 Ex connection data . . . . . . . . . . . . . . . . . . . . . . . . . . . 22 Low flow cut off . . . . . . . . . . . . . . . . . . . . . . . . . . . . . 23 Galvanic isolation . . . . . . . . . . . . . . . . . . . . . . . . . . . . 23 Protocol-specific data . . . . . . . . . . . . . . . . . . . . . . . . . . 24

Power supply . . . . . . . . . . . . . . . . . . . . . . . . . . . . .

28

Terminal assignment . . . . . . . . . . . . . . . . . . . . . . . . . .

28

Device plugs available . . . . . . . . . . . . . . . . . . . . . . . . . .

29

Pin assignment, device plug . . . . . . . . . . . . . . . . . . . . . .

30

Supply voltage . . . . . . . . . . . . . . . . . . . . . . . . . . . . . .

32

Power consumption . . . . . . . . . . . . . . . . . . . . . . . . . . .

32

Current consumption . . . . . . . . . . . . . . . . . . . . . . . . . .

32

Power supply failure . . . . . . . . . . . . . . . . . . . . . . . . . .

32

Electrical connection . . . . . . . . . . . . . . . . . . . . . . . . . .

33

Potential equalization . . . . . . . . . . . . . . . . . . . . . . . . .

43

terminals . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . .

44

Cable entries . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . .

44

Cable specification . . . . . . . . . . . . . . . . . . . . . . . . . . . .

44

Performance characteristics . . . . . . . . . . . . . . . . . . 49

Reference operating conditions . . . . . . . . . . . . . . . . . . . 49 Maximum measured error . . . . . . . . . . . . . . . . . . . . . . . 49 Repeatability . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . 51 Response time . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . 52 Influence of ambient temperature . . . . . . . . . . . . . . . . . 52 Influence of medium temperature . . . . . . . . . . . . . . . . . . 52 Influence of medium pressure . . . . . . . . . . . . . . . . . . . . 53 Design fundamentals . . . . . . . . . . . . . . . . . . . . . . . . . . 53

Installation . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . 54

Mounting location . . . . . . . . . . . . . . . . . . . . . . . . . . . . 54 Orientation . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . 55 Inlet and outlet runs . . . . . . . . . . . . . . . . . . . . . . . . . . 56 Mounting the transmitter housing . . . . . . . . . . . . . . . . . 57 Special mounting instructions . . . . . . . . . . . . . . . . . . . . 58

Environment . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . 60

Ambient temperature range . . . . . . . . . . . . . . . . . . . . . 60

Storage temperature . . . . . . . . . . . . . . . . . . . . . . . . . . 60 Climate class . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . 60 Degree of protection . . . . . . . . . . . . . . . . . . . . . . . . . . 60 Vibration resistance . . . . . . . . . . . . . . . . . . . . . . . . . . . 60 Shock resistance . . . . . . . . . . . . . . . . . . . . . . . . . . . . . 61 Shock resistance . . . . . . . . . . . . . . . . . . . . . . . . . . . . . 61 Interior cleaning . . . . . . . . . . . . . . . . . . . . . . . . . . . . . 61 Electromagnetic compatibility (EMC) . . . . . . . . . . . . . . . 61

Process . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . 61

Medium temperature range . . . . . . . . . . . . . . . . . . . . . . 61 Density . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . 62 Pressure-temperature curves . . . . . . . . . . . . . . . . . . . . . 63 Sensor housing . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . 68 Rupture disk . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . 69 Flow limit . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . 69 Pressure loss . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . 69 System pressure . . . . . . . . . . . . . . . . . . . . . . . . . . . . . 69 Thermal insulation . . . . . . . . . . . . . . . . . . . . . . . . . . . 70 Heating . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . 70 Vibrations . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . 71

Custody transfer measurement . . . . . . . . . . . . . . . 72

Mechanical construction . . . . . . . . . . . . . . . . . . . . 73

Dimensions in SI units . . . . . . . . . . . . . . . . . . . . . . . . . 73 Dimensions in US units . . . . . . . . . . . . . . . . . . . . . . . . . 96 Weight . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . 108 Materials . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . 109 Process connections . . . . . . . . . . . . . . . . . . . . . . . . . . 112 Surface roughness . . . . . . . . . . . . . . . . . . . . . . . . . . . 112

Operability . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . 112

Operating concept . . . . . . . . . . . . . . . . . . . . . . . . . . . 112 Languages . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . 113 Local operation . . . . . . . . . . . . . . . . . . . . . . . . . . . . . 113 Remote operation . . . . . . . . . . . . . . . . . . . . . . . . . . . 113 Service interface . . . . . . . . . . . . . . . . . . . . . . . . . . . . 119 Network integration . . . . . . . . . . . . . . . . . . . . . . . . . . 120 Supported operating tools . . . . . . . . . . . . . . . . . . . . . . 121 HistoROM data management . . . . . . . . . . . . . . . . . . . . 122

Certificates and approvals . . . . . . . . . . . . . . . . . . 124

CE mark . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . 124 C-Tick symbol . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . 124 Ex approval . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . 124 Sanitary compatibility . . . . . . . . . . . . . . . . . . . . . . . . . 127 Pharmaceutical compatibility . . . . . . . . . . . . . . . . . . . . 127 Functional safety . . . . . . . . . . . . . . . . . . . . . . . . . . . . 127 HART certification . . . . . . . . . . . . . . . . . . . . . . . . . . . 127 FOUNDATION Fieldbus certification . . . . . . . . . . . . . . . 127 Certification PROFIBUS . . . . . . . . . . . . . . . . . . . . . . . . 127 EtherNet/IP certification . . . . . . . . . . . . . . . . . . . . . . . 128 Certification PROFINET . . . . . . . . . . . . . . . . . . . . . . . . 128 Pressure Equipment Directive . . . . . . . . . . . . . . . . . . . 128 Radio approval . . . . . . . . . . . . . . . . . . . . . . . . . . . . . 128 Measuring instrument approval . . . . . . . . . . . . . . . . . . 128 Additional certification . . . . . . . . . . . . . . . . . . . . . . . . 128

2

Endress+Hauser

Proline Promass F 500

Other standards and guidelines . . . . . . . . . . . . . . . . . . 129

Ordering information . . . . . . . . . . . . . . . . . . . . . . 130

Application packages . . . . . . . . . . . . . . . . . . . . . . 130

Diagnostics functions . . . . . . . . . . . . . . . . . . . . . . . . . 130 Heartbeat Technology . . . . . . . . . . . . . . . . . . . . . . . . 130 Concentration . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . 131 Special density . . . . . . . . . . . . . . . . . . . . . . . . . . . . . 131 Petroleum . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . 131 OPC-UA server . . . . . . . . . . . . . . . . . . . . . . . . . . . . . 131

Accessories . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . 131

Device-specific accessories . . . . . . . . . . . . . . . . . . . . . . 132 Communication-specific accessories . . . . . . . . . . . . . . . 133 Service-specific accessories . . . . . . . . . . . . . . . . . . . . . 134 System components . . . . . . . . . . . . . . . . . . . . . . . . . . 134

Supplementary documentation . . . . . . . . . . . . . . 135

Standard documentation . . . . . . . . . . . . . . . . . . . . . . . 135 Device-dependent additional documentation . . . . . . . . . 136

Registered trademarks . . . . . . . . . . . . . . . . . . . . . 136

Endress+Hauser

3

Proline Promass F 500

About this document

Symbols used

Electrical symbols

 

 

 

 

 

 

 

Symbol

Meaning

 

 

 

 

 

 

 

 

 

Direct current

 

 

 

 

 

 

 

 

 

Alternating current

 

 

 

 

 

 

 

 

 

Direct current and alternating current

 

 

 

 

 

 

 

 

 

 

 

 

 

Ground connection

 

 

 

 

A grounded terminal which, as far as the operator is concerned, is grounded via a

 

 

 

 

grounding system.

 

 

 

 

 

 

 

 

 

Protective Earth (PE)

 

 

 

 

A terminal which must be connected to ground prior to establishing any other

 

 

 

 

connections.

 

 

 

 

The ground terminals are situated inside and outside the device:

 

 

 

 

• Inner ground terminal: Connects the protectiv earth to the mains supply.

 

 

 

 

• Outer ground terminal: Connects the device to the plant grounding system.

 

 

 

 

 

Communication symbols

Symbol Meaning

Wireless Local Area Network (WLAN)

Communication via a wireless, local network.

LED

Light emitting diode is off.

LED

Light emitting diode is on.

LED

Light emitting diode is flashing.

Symbols for certain types of information

Symbol Meaning

Permitted

Procedures, processes or actions that are permitted.

Preferred

Procedures, processes or actions that are preferred.

Forbidden

Procedures, processes or actions that are forbidden.

Tip

Indicates additional information.

Reference to documentation.

A

Reference to page.

Reference to graphic.

Visual inspection.

4

Endress+Hauser

Proline Promass F 500

Symbols in graphics

 

Symbol

Meaning

 

 

 

 

 

 

 

 

1, 2, 3, ...

Item numbers

 

 

 

 

 

 

 

 

 

 

,

 

,

 

, …

Series of steps

 

1.

2.

3.

 

 

 

 

A, B, C, ...

Views

 

 

A-A, B-B, C-C, ...

Sections

-Hazardous area

.

Safe area (non-hazardous area)

 

 

 

 

Flow direction

Function and system design

Measuring principle

The measuring principle is based on the controlled generation of Coriolis forces. These forces are

 

always present in a system when both translational and rotational movements are superimposed.

 

Fc =

2 · ∆m (ν · ω)

 

Fc =

Coriolis force

 

∆m =

moving mass

 

ω =

rotational velocity

 

ν =

radial velocity in rotating or oscillating system

The amplitude of the Coriolis force depends on the moving mass ∆m, its velocity ν in the system and thus on the mass flow. Instead of a constant rotational velocity ω, the sensor uses oscillation.

In the sensor, two parallel measuring tubes containing flowing fluid oscillate in antiphase, acting like a tuning fork. The Coriolis forces produced at the measuring tubes cause a phase shift in the tube oscillations (see illustration):

At zero flow (when the fluid is at a standstill) the two tubes oscillate in phase (1).

Mass flow causes deceleration of the oscillation at the inlet of the tubes (2) and acceleration at the outlet (3).

 

 

 

 

 

 

1

2

3

 

A0028850

The phase difference (A-B) increases with increasing mass flow. Electrodynamic sensors register the tube oscillations at the inlet and outlet. System balance is ensured by the antiphase oscillation of the two measuring tubes. The measuring principle operates independently of temperature, pressure, viscosity, conductivity and flow profile.

Endress+Hauser

5

Proline Promass F 500

Density measurement

The measuring tube is continuously excited at its resonance frequency. A change in the mass and thus the density of the oscillating system (comprising measuring tube and fluid) results in a corresponding, automatic adjustment in the oscillation frequency. Resonance frequency is thus a function of medium density. The microprocessor utilizes this relationship to obtain a density signal.

Volume measurement

Together with the measured mass flow, this is used to calculate the volume flow.

Temperature measurement

The temperature of the measuring tube is determined in order to calculate the compensation factor due to temperature effects. This signal corresponds to the process temperature and is also available as an output signal.

6

Endress+Hauser

Endress+Hauser F 500 Specifications

Proline Promass F 500

Measuring system

The measuring system consists of a transmitter and a sensor. The transmitter and sensor are

 

mounted in physically separate locations. They are interconnected by connecting cables.

Transmitter

Two versions of the transmitter are available.

Proline 500 – digital

Proline 500

 

 

 

 

 

 

For use in applications not required to meet special requirements due to

For use in applications required to meet special requirements due to

ambient or operating conditions.

ambient or operating conditions.

 

 

 

 

 

 

 

 

 

 

 

 

1

1

 

A

B

2

2

3

3

A

Non-hazardous area or Zone 2; Class I, Division 2

 

Non-hazardous area or Zone 2; Class I, Division 2 or Zone 1; Class I,

B

Non-hazardous area or Zone 2; Class I, Division 2 or Zone 1; Class I,

 

Division 1

 

Division 1

1

Transmitter with integrated ISEM

1

Transmitter

2

Connecting cable: cable, separate

2

Connecting cable: cable, separate, standard

3

Sensor connection housing

3Sensor connection housing with integrated ISEM

Flexible and cost-effective separate installation.

Application examples for sensors without electronics:

• A standard cable can be used as the connecting cable.

• Strong vibrations at the sensor.

 

 

• Sensor in underground installations.

 

 

• Permanent immersion of sensor in water, IP68 ingress protection.

• Electronics in the transmitter housing, ISEM (intelligent sensor

• Electronics and ISEM (intelligent sensor electronics module) in the

 

electronics module) in the sensor connection housing

transmitter housing

Signal transmission: digital

• Signal transmission: analog

 

Order code for "Integrated ISEM electronics", option A "Sensor"

Order code for "Integrated ISEM electronics", option B "Transmitter"

 

 

Connecting cable (can be ordered in various lengths)→ 132

 

 

 

• Length:

• Length: max. 20 m (65 ft)

 

– Zone 2; Class I, Division 2: max. 300 m (1 000 ft)

• Cable with a common shield and individual shielded cores (3 pairs)

– Zone 1; Class I, Division 1: max. 150 m (500 ft)

Standard cable with common shield (pair-stranded)

Hazardous area

Use in: Zone 2; Class I, Division 2

Use in: Zone 1; Class I, Division 1 oder Zone 2; Class I, Division 2

Mixed installation is possible:

Sensor: Zone 1; Class I, Division 1

Transmitter: Zone 2; Class I, Division 2

Device versions and materials

Transmitter housing

Transmitter housing

 

Aluminum, coated: aluminum, AlSi10Mg, coated

 

Aluminum, coated: aluminum, AlSi10Mg, coated

 

Material: polycarbonate

 

Cast, stainless: cast, stainless steel, 1.4409 (CF3M) similar to 316L

Material of window in transmitter housing

Window material: glass

Aluminum, coated: glass

Polycarbonate: plastic

Configuration

External operation via 4-line, backlit, graphic local display with touch control and guided menus ("Make-it-run" wizards) for application-specific commissioning.

Via service interface or WLAN interface:

Operating tools (e.g. FieldCare, DeviceCare, SmartBlue app)

Web server (access via Web browser, e.g. Microsoft Internet Explorer, Microsoft Edge)

Endress+Hauser

7

Proline Promass F 500

Sensor connection housing

Different versions of the connection housing are available.

Order code for "Sensor connection housing", option A, "Aluminum, coated": Aluminum, AlSi10Mg, coated

This device version is only available in conjunction with the Proline 500 – digital transmitter.

Order code for "Sensor connection housing", option B, "Stainless":

• Hygienic version, stainless steel 1.4301 (304)

• Optional: order code for "Sensor feature", option CC "Hygienic version, for maximum corrosion resistance": stainless steel 1.4404 (316L)

Order code for "Sensor connection housing", option C, "Ultra-compact hygienic, stainless":

• Hygienic version, stainless steel 1.4301 (304)

Optional: order code for "Sensor feature", option CC "Hygienic version, for maximum corrosion resistance": stainless steel 1.4404 (316L)

This device version is only available in conjunction with the Proline 500 – digital transmitter.

Order code for "Sensor connection housing", option L, "Cast, stainless": 1.4409 (CF3M) similar to 316L

Sensor

PromassF

 

• Excellent performance across a wide range of applications

 

 

• Simultaneous measurement of flow, volume flow, density and

 

 

temperature (multivariable)

 

 

• Immune to process influences

 

 

• Materials:

 

 

– Sensor: stainless steel, 1.4301/1.4307 (304L); optional 1.4404

 

A0026713

(316/316L)

Measuring tubes: stainless steel, 1.4539 (904L); 1.4404 (316/316L); Alloy C22, 2.4602 (UNS N06022)

Process connections: stainless steel, 1.4404 (316/316L); 1.4301 (304); Alloy C22, 2.4602 (UNS N06022)

8

Endress+Hauser

Proline Promass F 500

 

Equipment architecture

 

1

 

5

 

2

3

 

 

4

6

 

 

3

7

 

 

A0027512

1 Possibilities for integrating measuring devices into a system

 

1Control system (e.g. PLC)

2Connecting cable (0/4 to 20 mA HART etc.)

3Fieldbus

4Segment coupler

5Non-hazardous area

6 Hazardous area: Zone 2; Class I, Division 2

7Hazardous area: Zone 1; Class I, Division 1

Safety

IT security

Our warranty is valid only if the device is installed and used as described in the Operating Instructions. The device is equipped with security mechanisms to protect it against any inadvertent changes to the settings.

IT security measures, which provide additional protection for the device and associated data transfer, must be implemented by the operators themselves in line with their security standards.

Device-specific IT security

The device offers a range of specific functions to support protective measures on the operator's side. These functions can be configured by the user and guarantee greater in-operation safety if used correctly. An overview of the most important functions is provided in the following section.

Function/interface

Factory setting

Recommendation

 

 

 

Write protection via hardware write

Not enabled.

On an individual basis following risk

protection switch → 10

 

assessment.

 

 

 

Access code

Not enabled

Assign a customized access code during

(also applies for Web server login or

(0000).

commissioning.

FieldCare connection) → 10

 

 

 

 

 

WLAN

Enabled.

On an individual basis following risk

(order option in display module)

 

assessment.

 

 

 

WLAN security mode

Enabled (WPA2-

Do not change.

 

PSK)

 

 

 

 

WLAN passphrase

Serial number

Assign a customized access code during

(password) → 10

 

commissioning.

 

 

 

WLAN mode

Access Point

On an individual basis following risk

 

 

assessment.

 

 

 

Web server→ 10

Enabled.

On an individual basis following risk

 

 

assessment.

 

 

 

CDI-RJ45 service interface → 11

On an individual basis following risk

 

 

assessment.

 

 

 

Endress+Hauser

9

Proline Promass F 500

Protecting access via hardware write protection

Write access to the device parameters via the local display, Web browser or operating tool (e.g. FieldCare, DeviceCare) can be disabled via a write protection switch (DIP switch on the motherboard). When hardware write protection is enabled, only read access to the parameters is possible.

Hardware write protection is disabled when the device is delivered.

Protecting access via a password

Different passwords are available to protect write access to the device parameters or access to the device via the WLAN interface.

User-specific access code

Protect write access to the device parameters via the local display, Web browser or operating tool (e.g. FieldCare, DeviceCare). Access authorization is clearly regulated through the use of a userspecific access code.

WLAN passphrase

The network key protects a connection between an operating unit (e.g. notebook or tablet) and the device via the WLAN interface which can be ordered as an option.

Infrastructure mode

When the device is operated in infrastructure mode, the WLAN passphrase corresponds to the WLAN passphrase configured on the operator side.

User-specific access code

Write access to the device parameters via the local display, Web browser or operating tool (e.g. FieldCare, DeviceCare) can be protected by the modifiable, user-specific access code.

WLAN passphrase: Operation as WLAN access point

A connection between an operating unit (e.g. notebook or tablet) and the device via the WLAN interface, which can be ordered as an optional extra, is protected by the network key. The WLAN authentication of the network key complies with the IEEE 802.11 standard.

When the device is delivered, the network key is pre-defined depending on the device. It can be changed via the WLAN settings submenu in the WLAN passphrase parameter.

Infrastructure mode

A connection between the device and WLAN access point is protected by means of an SSID and passphrase on the system side. Please contact the relevant system administrator for access.

General notes on the use of passwords

The access code and network key supplied with the device should be changed during commissioning.

Follow the general rules for generating a secure password when defining and managing the access code or network key.

The user is responsible for the management and careful handling of the access code and network key.

Access via Web server

The device can be operated and configured via a Web browser with the integrated Web server. The connection is via the service interface (CDI-RJ45) or the WLAN interface. For device versions with the EtherNet/IP and PROFINET communication protocols, the connection can also be established via the terminal connection for signal transmission with EtherNet/IP or PROFINET (RJ45 connector).

The Web server is enabled when the device is delivered. The Web server can be disabled if necessary (e.g. after commissioning) via the Web server functionality parameter.

The device and status information can be hidden on the login page. This prevents unauthorized access to the information.

For detailed information on device parameters, see:

The "Description of Device Parameters" document → 135

Access via OPC-UA

The "OPC UA Server" application package is available in the device version with the HART communication protocol → 131.

10

Endress+Hauser

Proline Promass F 500

The device can communicate with OPC UA clients using the "OPC UA Server" application package.

The OPC UA server integrated in the device can be accessed via the WLAN access point using the WLAN interface - which can be ordered as an optional extra - or the service interface (CDIRJ45) via Ethernet network. Access rights and authorization as per separate configuration.

The following Security Modes are supported as per the OPC UA Specification (IEC 62541):

None

Basic128Rsa15 – signed

Basic128Rsa15 – signed and encrypted

Access via service interface (CDI-RJ45)

The device can be connected to a network via the service interface (CDI-RJ45). Device-specific functions guarantee the secure operation of the device in a network.

The use of relevant industrial standards and guidelines that have been defined by national and international safety committees, such as IEC/ISA62443 or the IEEE, is recommended. This includes organizational security measures such as the assignment of access authorization as well as technical measures such as network segmentation.

The device can be integrated in a ring topology. The device is integrated via the terminal connection for signal transmission (output 1) and the connection to the service interface (CDIRJ45) → 117.

Endress+Hauser

11

Proline Promass F 500

 

Input

 

 

 

 

 

 

 

 

 

Measured variable

Direct measured variables

 

 

 

 

• Mass flow

 

 

 

 

• Density

 

 

 

 

• Temperature

 

 

 

 

Calculated measured variables

 

 

 

 

• Volume flow

 

 

 

 

• Corrected volume flow

 

 

 

 

• Reference density

 

 

 

 

 

 

 

 

Measuring range

Measuring ranges for liquids

 

 

 

 

 

 

 

 

 

 

DN

 

Measuring range full scale values min(F) to max(F)

 

[mm]

 

[in]

[kg/h]

[lb/min]

 

 

 

 

 

 

 

8

 

³⁄

0 to 2 000

0 to 73.50

 

 

 

 

 

 

 

15

 

½

0 to 6 500

0 to 238.9

 

 

 

 

 

 

 

25

 

1

0 to 18 000

0 to 661.5

 

 

 

 

 

 

 

40

 

0 to 45 000

0 to 1 654

 

 

 

 

 

 

 

50

 

2

0 to 70 000

0 to 2 573

 

 

 

 

 

 

 

80

 

3

0 to 180 000

0 to 6 615

 

 

 

 

 

 

 

100

 

4

0 to 350 000

0 to 12 860

 

 

 

 

 

 

 

150

 

6

0 to 800 000

0 to 29 400

 

 

 

 

 

 

 

250

 

10

0 to 2 200 000

0 to 80 850

 

 

 

 

 

 

Measuring range for gases

The full scale value depends on the density and the sound velocity of the gas used and can be calculated with the formula below:

max(G) = minimum ( max(F) · ρG : x ; ρG · cG · π/2 · (di)2 · 3600)

max(G)

Maximum full scale value for gas [kg/h]

 

max(F)

Maximum full scale value for liquid [kg/h]

 

max(G) < max(F)

max(G) can never be greater than max(F)

 

ρG

Gas density in [kg/m³] at operating conditions

 

 

 

 

 

 

x

Constant dependent on nominal diameter

 

 

 

 

 

 

cG

Sound velocity (gas) [m/s]

 

di

Measuring tube internal diameter [m]

 

 

 

 

 

 

DN

 

 

x

 

 

 

 

[mm]

 

[in]

 

[kg/m3]

8

 

³⁄

 

60

 

 

 

 

 

15

 

½

 

80

 

 

 

 

 

25

 

1

 

90

 

 

 

 

 

40

 

 

90

 

 

 

 

 

50

 

2

 

90

 

 

 

 

 

80

 

3

 

110

 

 

 

 

 

12

Endress+Hauser

Proline Promass F 500

 

DN

x

[mm]

 

[in]

[kg/m3]

 

100

 

4

130

 

 

 

 

150

 

6

200

 

 

 

 

250

 

10

200

 

 

 

 

To calculate the measuring range, use the Applicator sizing tool → 134

Calculation example for gas

Sensor: Promass F, DN 50

Gas: Air with a density of 60.3 kg/m³ (at 20 °C and 50 bar)

Measuring range (liquid): 70 000 kg/h

x = 90 kg/m³ (for Promass F, DN 50)

Maximum possible full scale value:

max(G) = max(F) · ρG : x = 70 000 kg/h · 60.3 kg/m³ : 90 kg/m³ = 46 900 kg/h

 

Recommended measuring range

 

"Flow limit" section → 69

 

 

Operable flow range

Over 1000 : 1.

 

Flow rates above the preset full scale value do not override the electronics unit, with the result that

 

the totalizer values are registered correctly.

 

 

Input signal

Input and output versions

 

→ 15

 

External measured values

 

To increase the accuracy of certain measured variables or to calculate the corrected volume flow for

 

gases, the automation system can continuously write different measured values to the measuring

 

device:

 

• Operating pressure to increase accuracy (Endress+Hauser recommends the use of a pressure

 

measuring device for absolute pressure, e.g. Cerabar M or Cerabar S)

 

• Medium temperature to increase accuracy (e.g. iTEMP)

 

• Reference density for calculating the corrected volume flow for gases

 

Various pressure transmitters and temperature measuring devices can be ordered from Endress

 

+Hauser: see "Accessories" section → 134

 

It is recommended to read in external measured values to calculate the corrected volume flow.

 

HART protocol

 

The measured values are written from the automation system to the measuring device via the HART

 

protocol. The pressure transmitter must support the following protocol-specific functions:

 

• HART protocol

 

• Burst mode

 

Current input

 

The measured values are written from the automation system to the measuring device via the

 

current input → 14.

 

Digital communication

 

The measured values can be written from the automation system to the measuring via:

 

• FOUNDATION Fieldbus

 

• PROFIBUS DP

 

• PROFIBUS PA

 

• Modbus RS485

 

• EtherNet/IP

 

• PROFINET

Endress+Hauser

13

Proline Promass F 500

Current input 0/4 to 20 mA

Current input

0/4 to 20 mA (active/passive)

 

 

Current span

• 4 to 20 mA (active)

 

• 0/4 to 20 mA (passive)

 

 

 

Resolution

1

µA

 

 

Voltage drop

Typically: 0.6 to 2 V for 3.6 to 22 mA (passive)

 

 

Maximum input voltage

≤ 30 V (passive)

 

 

Open-circuit voltage

≤ 28.8 V (active)

 

 

Possible input variables

• Pressure

 

• Temperature

 

• Density

 

 

 

Status input

Maximum input values

• DC –3 to 30 V

 

• If status input is active (ON): Ri >3 kΩ

Response time

Adjustable: 5 to 200 ms

 

 

Input signal level

• Low signal: DC –3 to +5 V

 

• High signal: DC 12 to 30 V

 

 

Assignable functions

• Off

 

Reset the individual totalizers separately

 

Reset all totalizers

 

• Flow override

 

 

 

14

Endress+Hauser

Proline Promass F 500

 

Output

 

 

 

 

 

 

 

 

 

 

 

 

Output and input variants

Depending on the option selected for output/input 1, different options are available for the other

 

outputs and inputs. Only one option can be selected for each output/input 1 to 4. The table must be

 

read vertically (↓).

 

 

 

 

 

 

 

 

 

 

 

Example: If the option BA "4–20 mA HART" was selected for output/input 1, one of the options A, B,

 

D, E, F, H, I or J is available for output 2, and one of the options A, B, D, E, F, H, I or J is available for

 

output 3 and 4.

 

 

 

 

 

 

 

 

 

 

 

 

 

 

 

 

 

 

 

 

 

Order code for "Output; input 1" (020) →

 

 

 

 

Possible options

 

 

 

 

 

 

 

 

 

 

 

 

 

 

 

Current output 4 to 20 mA HART

 

BA

 

 

 

 

 

 

 

 

 

 

 

 

 

 

 

 

 

 

 

 

Current output 4 to 20 mA HART Ex i

CA

 

 

 

 

 

 

 

 

 

 

 

 

 

 

 

 

 

 

 

 

FOUNDATION Fieldbus

 

 

SA

 

 

 

 

 

 

 

 

 

 

 

 

 

 

 

 

 

 

 

FOUNDATION Fieldbus Ex i

 

 

 

TA

 

 

 

 

 

 

 

 

 

 

 

 

 

 

 

 

 

 

PROFIBUS DP

 

 

 

 

LA

 

 

 

 

 

 

 

 

 

 

 

 

 

 

 

 

 

PROFIBUS PA

 

 

 

 

 

GA

 

 

 

 

 

 

 

 

 

 

 

 

 

 

 

 

PROFIBUS PA Ex i

 

 

 

 

 

 

HA

 

 

 

 

 

 

 

 

 

 

 

 

 

 

 

Modbus RS485

 

 

 

 

 

 

 

MA

 

 

 

 

 

 

 

 

 

 

 

 

 

 

EtherNet/IP 2-port switch integrated

 

 

 

 

 

 

 

 

NA

 

 

 

 

 

 

 

 

 

 

 

 

 

PROFINET 2-port switch integrated

 

 

 

 

 

 

 

 

 

RA

 

 

 

 

 

 

 

 

 

 

 

Order code for "Output; input 2" (021) →

 

 

 

 

 

 

 

 

 

 

 

 

Not assigned

 

A

A

A

A

A

A

A

A

A

A

 

 

 

 

 

 

 

 

 

 

 

 

Current output 0/4 to 20 mA

 

B

 

B

 

B

B

 

B

B

B

 

 

 

 

 

 

 

 

 

 

 

 

Current output 0/4 to 20 mA (Ex i)

 

 

C

 

C

 

 

C

 

 

 

 

 

 

 

 

 

 

 

 

 

 

 

User configurable input/output 1)

 

D

 

D

 

D

D

 

D

D

D

Pulse/frequency/switch output

 

E

 

E

 

E

E

 

E

E

E

 

 

 

 

 

 

 

 

 

 

 

 

Double pulse output 2)

 

F

 

 

 

 

 

 

F

 

 

Pulse/frequency/switch output (Ex i)

 

 

G

 

G

 

 

G

 

 

 

 

 

 

 

 

 

 

 

 

 

 

 

Relay output

 

H

 

H

 

H

H

 

H

H

H

 

 

 

 

 

 

 

 

 

 

 

 

Current input 0/4 to 20 mA

 

I

 

I

 

I

I

 

I

I

I

 

 

 

 

 

 

 

 

 

 

 

 

Status input

 

J

 

J

 

J

J

 

J

J

J

 

 

 

 

 

 

 

 

 

 

 

Order code for "Output; input 3" (022), "Output; input 4" (023) 3)

Not assigned

 

A

A

A

A

A

A

A

A

A

A

 

 

 

 

 

 

 

 

 

 

 

 

Current output 0/4 to 20 mA

 

B

 

 

 

B

 

 

B

B

B

 

 

 

 

 

 

 

 

 

 

 

 

Current output 0/4 to 20 mA (Ex i)

 

 

C

 

 

 

 

 

 

 

 

 

 

 

 

 

 

 

 

 

 

 

 

User configurable input/output

 

D

 

 

 

D

 

 

D

D

D

 

 

 

 

 

 

 

 

 

 

 

 

Pulse/frequency/switch output

 

E

 

 

 

E

 

 

E

E

E

 

 

 

 

 

 

 

 

 

 

 

 

Double pulse output (slave) 4)

 

F

 

 

 

 

 

 

F

 

 

Pulse/frequency/switch output (Ex i)

 

 

G

 

 

 

 

 

 

 

 

 

 

 

 

 

 

 

 

 

 

 

 

Relay output

 

H

 

 

 

H

 

 

H

H

H

 

 

 

 

 

 

 

 

 

 

 

 

Current input 0/4 to 20 mA

 

I

 

 

 

I

 

 

I

I

I

 

 

 

 

 

 

 

 

 

 

 

 

Status input

 

J

 

 

 

J

 

 

J

J

J

 

 

 

 

 

 

 

 

 

 

 

 

1)A specific input or output can be assigned to a user configurable input/output → 19.

2)If double pulse output (F) is selected for output/input 2 (021), only the double pulse output (F) option is available for selection for output/input 3 (022).

3)The order code for "Output; input 4" (023) is only available for the Proline 500 – digital transmitter.

4)The double pulse output (F) option is not available for input/output 4.

Endress+Hauser

15

Proline Promass F 500

Output signal

HART current output

 

 

 

 

 

Current output

4 to 20 mA HART

 

 

 

 

Current span

Can be set to: 4 to 20 mA (active/passive)

 

 

Ex-i, passive

 

Open-circuit voltage

DC 28.8 V (active)

 

 

 

 

Maximum input voltage

DC 30 V (passive)

 

 

 

 

Load

250 to 700 Ω

 

 

 

 

Resolution

0.38 µA

 

 

 

 

Damping

Configurable: 0.07 to 999 s

 

 

 

 

Assignable measured

• Mass flow

 

variables

• Volume flow

 

 

• Corrected volume flow

 

 

• Density

 

 

• Reference density

 

 

• Temperature

 

 

• Electronic temperature

 

 

• Oscillation frequency 0

 

 

• Oscillation damping 0

 

 

• Signal asymmetry

 

 

• Exciter current 0

 

 

The range of options increases if the measuring device has one or more

 

 

application packages.

 

PROFIBUS PA

 

 

 

 

 

PROFIBUS PA

In accordance with EN 50170 Volume 2, IEC 61158-2 (MBP), galvanically

 

 

isolated

 

 

 

 

Data transmission

31.25 kbit/s

 

 

 

 

Current consumption

10 mA

 

 

16 mA

 

 

 

 

Permitted supply voltage

9 to 32 V

 

 

 

 

Bus connection

With integrated reverse polarity protection

 

 

 

 

PROFIBUS DP

 

 

 

 

 

Signal encoding

NRZ code

 

 

 

 

Data transfer

9.6 kBaud…12 MBaud

 

 

 

 

EtherNet/IP

 

 

 

 

 

Standards

In accordance with IEEE 802.3

 

 

 

 

PROFINET

 

 

 

 

 

Standards

In accordance with IEEE 802.3

 

 

 

 

FOUNDATION Fieldbus

 

 

 

 

 

FOUNDATION Fieldbus

H1, IEC 61158-2, galvanically isolated

 

 

 

 

Data transfer

31.25 kbit/s

 

 

 

16

Endress+Hauser

Proline Promass F 500

Current consumption

10 mA

 

 

Permitted supply voltage

9 to 32 V

 

 

Bus connection

With integrated reverse polarity protection

 

 

Modbus RS485

Physical interface

RS485 in accordance with EIA/TIA-485 standard

 

 

Terminating resistor

Integrated, can be activated via DIP switches

 

 

Current output 0/4 to 20 mA

Current output

0/4 to 20 mA

 

 

Maximum output values

22.5 mA

 

 

Current span

Can be set to:

 

• 4 to 20 mA (active)

 

• 0/4 to 20 mA (passive)

 

Ex-i, passive

Open-circuit voltage

DC 28.8 V (active)

 

 

Maximum input voltage

DC 30 V (passive)

 

 

Load

0 to 700 Ω

 

 

Resolution

0.38 µA

 

 

Damping

Adjustable: 0.07 to 999 s

 

 

Assignable measured

• Mass flow

variables

• Volume flow

 

• Corrected volume flow

 

• Density

 

• Reference density

 

• Temperature

 

• Electronic temperature

 

• Oscillation frequency 0

 

• Oscillation damping 0

 

• Signal asymmetry

 

• Exciter current 0

 

The range of options increases if the measuring device has one or more

 

application packages.

Pulse/frequency/switch output

 

 

Function

Can be set to pulse, frequency or switch output

 

 

Version

Open collector

 

Can be set to:

 

• Active

 

• Passive

 

Ex-i, passive

Maximum input values

DC 30 V, 250 mA (passive)

 

 

Open-circuit voltage

DC 28.8 V (active)

 

 

Voltage drop

For 22.5 mA: ≤ DC 2 V

 

 

Pulse output

 

 

 

Maximum input values

DC 30 V, 250 mA (passive)

 

 

Maximum output current

22.5 mA (active)

 

 

Endress+Hauser

17

Proline Promass F 500

Open-circuit voltage

DC 28.8 V (active)

 

 

Pulse width

Adjustable: 0.05 to 2 000 ms

 

 

Maximum pulse rate

10 000 Impulse/s

 

 

Pulse value

Adjustable

 

 

Assignable measured

• Mass flow

variables

• Volume flow

 

• Corrected volume flow

 

 

Frequency output

 

 

 

Maximum input values

DC 30 V, 250 mA (passive)

 

 

Maximum output current

22.5 mA (active)

 

 

Open-circuit voltage

DC 28.8 V (active)

 

 

Output frequency

Adjustable: end value frequency 2 to 10 000 Hz (f max = 12 500 Hz)

Damping

Adjustable: 0 to 999 s

 

 

Pulse/pause ratio

1:1

 

 

Assignable measured

• Mass flow

variables

• Volume flow

 

• Corrected volume flow

 

• Density

 

• Reference density

 

• Temperature

 

• Electronic temperature

 

• Oscillation frequency 0

 

• Oscillation damping 0

 

• Signal asymmetry

 

• Exciter current 0

 

The range of options increases if the measuring device has one or more

 

application packages.

Switch output

 

 

 

Maximum input values

DC 30 V, 250 mA (passive)

 

 

Open-circuit voltage

DC 28.8 V (active)

 

 

Switching behavior

Binary, conductive or non-conductive

 

 

Switching delay

Adjustable: 0 to 100 s

 

 

Number of switching

Unlimited

cycles

 

 

 

Assignable functions

• Off

 

• On

 

• Diagnostic behavior

 

• Limit value

 

– Mass flow

 

– Volume flow

 

– Corrected volume flow

 

– Density

 

– Reference density

 

– Temperature

 

– Totalizer 1-3

 

• Flow direction monitoring

 

• Status

 

– Partially filled pipe detection

 

– Low flow

 

The range of options increases if the measuring device has one or more

 

application packages.

18

Endress+Hauser

Proline Promass F 500

Double pulse output

Function

Double pulse

 

 

Version

Open collector

 

Can be set to:

 

• Active

 

• Passive

 

 

Maximum input values

DC 30 V, 250 mA (passive)

 

 

Open-circuit voltage

DC 28.8 V (active)

 

 

Voltage drop

For 22.5 mA: ≤ DC 2 V

 

 

Output frequency

Adjustable: 0 to 1 000 Hz

 

 

Damping

Adjustable: 0 to 999 s

 

 

Pulse/pause ratio

1:1

 

 

Assignable measured

• Mass flow

variables

• Volume flow

 

• Corrected volume flow

 

• Density

 

• Reference density

 

• Temperature

 

The range of options increases if the measuring device has one or more

 

application packages.

Relay output

 

 

 

Function

Switch output

 

 

Version

Relay output, galvanically isolated

 

 

Switching behavior

Can be set to:

 

• NO (normally open), factory setting

 

• NC (normally closed)

 

 

Maximum switching

• DC 30 V, 0.1 A

capacity (passive)

• AC 30 V, 0.5 A

 

 

Assignable functions

• Off

 

• On

 

• Diagnostic behavior

 

• Limit value

 

– Mass flow

 

– Volume flow

 

– Corrected volume flow

 

– Density

 

– Reference density

 

– Temperature

 

– Totalizer 1-3

 

• Flow direction monitoring

 

• Status

 

– Partially filled pipe detection

 

– Low flow

 

The range of options increases if the measuring device has one or more

 

application packages.

User configurable input/output

One specific input or output is assigned to a user-configurable input/output (configurable I/O) during device commissioning.

The following inputs and outputs are available for assignment:

Choice of current output: 4 to 20 mA (active), 0/4 to 20 mA (passive)

Pulse/frequency/switch output

Choice of current input: 4 to 20 mA (active), 0/4 to 20 mA (passive)

Status input

Endress+Hauser

19

 

 

Proline Promass F 500

 

 

 

 

The technical values correspond to those of the inputs and outputs described in this section.

 

 

Signal on alarm

Depending on the interface, failure information is displayed as follows:

 

HART current output

 

 

 

 

 

Device diagnostics

Device condition can be read out via HART Command 48

 

 

 

 

PROFIBUS PA

 

 

 

 

 

Status and alarm

Diagnostics in accordance with PROFIBUS PA Profile 3.02

 

messages

 

 

 

 

 

Error current FDE (Fault

0 mA

 

Disconnection Electronic)

 

 

 

 

 

PROFIBUS DP

 

 

 

 

 

Status and alarm

Diagnostics in accordance with PROFIBUS PA Profile 3.02

 

messages

 

 

 

 

 

EtherNet/IP

 

 

 

 

 

Device diagnostics

Device condition can be read out in Input Assembly

 

 

 

 

PROFINET

 

 

 

 

 

Device diagnostics

According to "Application Layer protocol for decentralized periphery", Version 2.3

 

 

 

 

FOUNDATION Fieldbus

 

 

 

 

 

Status and alarm

Diagnostics in accordance with FF-891

 

messages

 

 

 

 

 

Error current FDE (Fault

0 mA

 

Disconnection Electronic)

 

 

 

 

 

Modbus RS485

 

 

 

 

 

Failure mode

Choose from:

 

 

• NaN value instead of current value

 

 

• Last valid value

 

 

 

Current output 0/4 to 20 mA

4 to 20 mA

Failure mode

Choose from:

 

• 4 to 20 mA in accordance with NAMUR recommendation NE 43

 

• 4 to 20 mA in accordance with US

 

• Min. value: 3.59 mA

 

• Max. value: 22.5 mA

 

• Freely definable value between: 3.59 to 22.5 mA

 

• Actual value

 

• Last valid value

 

 

20

Endress+Hauser

Proline Promass F 500

0 to 20 mA

Failure mode

Choose from:

 

• Maximum alarm: 22 mA

 

• Freely definable value between: 0 to 20.5 mA

 

 

Pulse/frequency/switch output

Pulse output

Failure mode

Choose from:

 

• Actual value

 

• No pulses

 

 

Frequency output

 

 

 

Failure mode

Choose from:

 

• Actual value

 

• 0 Hz

 

• Defined value (f max 2 to 12 500 Hz)

Switch output

 

 

 

Failure mode

Choose from:

 

• Current status

 

• Open

 

• Closed

 

 

Relay output

Failure mode

Choose from:

 

• Current status

 

• Open

 

• Closed

 

 

Local display

Plain text display

With information on cause and remedial measures

 

 

Backlight

Red backlighting indicates a device error.

 

 

Status signal as per NAMUR recommendation NE 107

Interface/protocol

Via digital communication:

HART protocol

FOUNDATION Fieldbus

PROFIBUS PA

PROFIBUS DP

Modbus RS485

EtherNet/IP

PROFINET

Via service interface

CDI-RJ45 service interface

WLAN interface

Plain text display

With information on cause and remedial measures

 

 

Additional information on remote operation → 113

Endress+Hauser

21

Proline Promass F 500

Web server

Plain text display

With information on cause and remedial measures

 

 

Light emitting diodes (LED)

Status information

Status indicated by various light emitting diodes

The following information is displayed depending on the device version:

Supply voltage active

Data transmission active

Device alarm/error has occurred

EtherNet/IP network available

EtherNet/IP connection established

PROFINET network available

PROFINET connection established

PROFINET blinking feature

Ex connection data

Safety-related values

 

 

 

 

 

 

 

 

 

 

 

 

 

 

 

 

 

 

 

 

 

 

 

 

 

 

Order code for

 

Output type

 

 

 

 

Safety-related values

 

"Output; input 1"

 

 

 

 

 

 

"Output; input 1"

 

 

 

 

 

 

 

26 (+)

 

 

 

27 (–)

 

 

 

 

 

 

 

 

 

 

 

 

 

 

 

 

 

 

 

 

 

 

 

Option BA

 

Current output

 

UN = 30 VDC

 

 

 

 

 

 

 

 

 

 

4 to 20 mA HART

 

UM = 250 VAC

 

 

 

 

 

 

 

 

Option GA

 

PROFIBUS PA

 

UN = 30 VDC

 

 

 

 

 

 

 

 

 

 

 

 

UM = 250 VAC

 

 

 

 

 

 

 

 

Option LA

 

PROFIBUS DP

 

UN = 30 VDC

 

 

 

 

 

 

 

 

 

 

 

 

UM = 250 VAC

 

 

 

 

 

 

 

 

Option MA

 

Modbus RS485

 

UN = 30 VDC

 

 

 

 

 

 

 

 

 

 

 

 

UM = 250 VAC

 

 

 

 

 

 

 

 

Option SA

 

FOUNDATION Fieldbus

 

UN = 30 VDC

 

 

 

 

 

 

 

 

 

 

 

 

UM = 250 VAC

 

 

 

 

 

 

 

 

Option NA

 

EtherNet/IP

 

UN = 30 VDC

 

 

 

 

 

 

 

 

 

 

 

 

UM = 250 VAC

 

 

 

 

 

 

 

 

Option RA

 

PROFINET

 

UN = 30 VDC

 

 

 

 

 

 

 

 

 

 

 

 

UM = 250 VAC

 

 

 

 

 

 

 

 

 

 

 

 

 

 

 

 

 

 

 

 

Order code for

Output type

 

 

 

Safety-related values

 

 

 

"Output; input 2";

 

 

Output; input 2

 

Output; input 3

 

Output; input

 

 

 

 

 

 

"Output; input 3"

 

 

 

 

 

 

 

 

 

 

 

 

 

 

 

4 1)

 

"Output; input 4"

 

 

 

 

 

 

 

 

 

 

 

 

 

 

24 (+)

25 (–)

 

22 (+)

23 (–)

 

20 (+)

21 (–)

 

 

 

 

 

 

 

 

 

 

 

 

 

Option B

Current output

UN = 30 VDC

 

 

 

 

 

 

 

 

 

4 to 20 mA

UM = 250 VAC

 

 

 

 

 

 

 

 

Option D

User configurable input/

UN = 30 VDC

 

 

 

 

 

 

 

 

 

output

UM = 250 VAC

 

 

 

 

 

 

 

 

Option E

Pulse/frequency/switch

UN = 30 VDC

 

 

 

 

 

 

 

 

 

output

UM = 250 VAC

 

 

 

 

 

 

 

 

Option F

Double pulse output

UN = 30 VDC

 

 

 

 

 

 

 

 

 

 

 

UM = 250 VAC

 

 

 

 

 

 

 

 

Option H

Relay output

UN = 30 VDC

 

 

 

 

 

 

 

 

 

 

 

IN =100 mADC/500 mAAC

 

 

 

 

 

 

UM = 250 VAC

 

 

 

 

 

 

 

22

Endress+Hauser

Proline Promass F 500

Order code for

Output type

 

 

Safety-related values

 

 

"Output; input 2";

 

Output; input 2

 

Output; input 3

 

Output; input

 

 

 

"Output; input 3"

 

 

 

 

 

 

 

 

 

 

4 1)

"Output; input 4"

 

 

 

 

 

 

 

 

 

24 (+)

25 (–)

 

22 (+)

23 (–)

 

20 (+)

21 (–)

 

 

 

 

 

 

 

 

 

 

Option I

Current input

UN = 30 VDC

 

 

 

 

 

 

 

4 to 20 mA

UM = 250 VAC

 

 

 

 

 

 

Option J

Status input

UN = 30 VDC

 

 

 

 

 

 

 

 

UM = 250 VAC

 

 

 

 

 

 

1)The order code "Output; input 4" is only available for the Proline 500 – digital transmitter.

Intrinsically safe values

Order code for

Output type

Intrinsically safe values

"Output; input 1"

 

 

"Output; input 1"

 

 

26 (+)

 

27 (–)

 

 

 

 

 

 

 

 

Option CA

Current output

Ui = 30 V

 

 

 

4 to 20 mA HART Ex i

li = 100 mA

 

 

 

 

Pi = 1.25 W

 

 

 

 

Li = 0

 

 

 

 

Ci = 0

 

 

Option HA

PROFIBUS PA Ex i

Ex ia 1)

 

Ex ic 2)

 

 

Ui = 30 V

 

Ui = 32 V

 

 

li = 570 mA

 

li = 570 mA

 

 

Pi = 8.5 W

 

Pi = 8.5 W

 

 

Li = 10 µH

 

Li = 10 µH

 

 

Ci = 5 nF

 

Ci = 5 nF

Option TA

FOUNDATION Fieldbus

Ex ia 1)

 

Ex ic 2)

 

Ex i

Ui = 30 V

 

Ui = 32 V

 

 

li = 570 mA

 

li = 570 mA

 

 

Pi = 8.5 W

 

Pi = 8.5 W

 

 

Li = 10 µH

 

Li = 10 µH

 

 

Ci = 5 nF

 

Ci = 5 nF

1)Only available for the Zone 1; Class I, Division 1 version

2)Only available for the Zone 2; Class I, Division 2 version and only for the Proline 500 – digital transmitter

Order code for

Output type

Intrinsically safe values or NIFW values

"Output; input 2";

 

Output; input 2

 

Output; input 3

 

Output; input 4 1)

 

 

 

"Output; input 3"

 

 

 

"Output; input 4"

 

24 (+)

25 (–)

 

22 (+)

23 (–)

 

20 (+)

21 (–)

 

 

 

 

 

 

 

 

 

 

Option C

Current output

Ui = 30 V

 

 

 

 

 

 

 

4 to 20 mA Ex i

li = 100 mA

 

 

 

 

 

 

 

 

Pi = 1.25 W

 

 

 

 

 

 

 

 

Li = 0

 

 

 

 

 

 

 

 

Ci = 0

 

 

 

 

 

 

Option G

Pulse/frequency/switch

Ui = 30 V

 

 

 

 

 

 

 

output Ex i

li = 100 mA

 

 

 

 

 

 

 

 

Pi = 1.25 W

 

 

 

 

 

 

 

 

Li = 0

 

 

 

 

 

 

 

 

Ci = 0

 

 

 

 

 

 

1)The order code "Output; input 4" is only available for the Proline 500 – digital transmitter.

Low flow cut off

The switch points for low flow cut off are user-selectable.

 

 

Galvanic isolation

The outputs are galvanically isolated from one another and from earth (PE).

Endress+Hauser

23

Proline Promass F 500

Protocol-specific data

HART

 

 

 

 

 

Manufacturer ID

0x11

 

 

 

 

Device type ID

0x3B

 

 

 

 

HART protocol revision

7

 

 

 

 

Device description files

Information and files under:

 

(DTM, DD)

www.endress.com

 

 

 

 

HART load

Min. 250 Ω

 

 

 

 

System integration

Information on system integration: Operating Instructions → 135.

 

 

• Measured variables via HART protocol

 

 

• Burst Mode functionality

 

 

 

PROFIBUS PA

Manufacturer ID

0x11

 

 

Ident number

0x156D

 

 

Profile version

3.02

 

 

Device description files (GSD,

Information and files under:

DTM, DD)

www.endress.com

 

www.profibus.org

 

 

Supported functions

• Identification & Maintenance

 

Simplest device identification on the part of the control system and

 

nameplate

 

• PROFIBUS upload/download

 

Reading and writing parameters is up to ten times faster with PROFIBUS

 

upload/download

 

• Condensed status

 

Simplest and self-explanatory diagnostic information by categorizing

 

diagnostic messages that occur

 

 

Configuration of the device

• DIP switches on the I/O electronics module

address

• Local display

 

• Via operating tools (e.g. FieldCare)

 

 

Compatibility with

If the device is replaced, the measuring device Promass 500 supports the

earlier model

compatibility of the cyclic data with previous models. It is not necessary to

 

adjust the engineering parameters of the PROFIBUS network with the

 

Promass 500 GSD file.

 

Earlier models:

 

• Promass 80 PROFIBUS PA

 

– ID No.: 1528 (hex)

 

– Extended GSD file: EH3x1528.gsd

 

– Standard GSD file: EH3_1528.gsd

 

• Promass 83 PROFIBUS PA

 

– ID No.: 152A (hex)

 

– Extended GSD file: EH3x152A.gsd

 

– Standard GSD file: EH3_152A.gsd

 

Description of the function scope of compatibility:

 

Operating Instructions → 135.

System integration

Information regarding system integration: Operating Instructions → 135.

 

• Cyclic data transmission

 

• Block model

 

• Description of the modules

 

 

PROFIBUS DP

 

 

 

Manufacturer ID

0x11

 

 

Ident number

0x156F

 

 

24

Endress+Hauser

Proline Promass F 500

Profile version

3.02

 

 

Device description files (GSD,

Information and files under:

DTM, DD)

www.endress.com

 

On the product page for the device: Documents/Software → Device drivers

 

www.profibus.org

 

 

Supported functions

• Identification & Maintenance

 

Simplest device identification on the part of the control system and

 

nameplate

 

• PROFIBUS upload/download

 

Reading and writing parameters is up to ten times faster with PROFIBUS

 

upload/download

 

• Condensed status

 

Simplest and self-explanatory diagnostic information by categorizing

 

diagnostic messages that occur

 

 

Configuration of the device

• DIP switches on the I/O electronics module

address

• Via operating tools (e.g. FieldCare)

 

 

System integration

Information regarding system integration: Operating Instructions → 135.

 

• Cyclic data transmission

 

• Block model

 

• Description of the modules

 

 

EtherNet/IP

Protocol

• The CIP Networks Library Volume 1: Common Industrial Protocol

 

• The CIP Networks Library Volume 2: EtherNet/IP Adaptation of CIP

 

 

Communication type

• 10Base-T

 

• 100Base-TX

 

 

Device profile

Generic device (product type: 0x2B)

 

 

Manufacturer ID

0x11

 

 

Device type ID

0x103B

 

 

Baud rates

Automatic ¹ ⁄ Mbit with half-duplex and full-duplex detection

 

 

Polarity

Auto-polarity for automatic correction of crossed TxD and RxD pairs

 

 

Supported CIP connections

Max. 3 connections

 

 

Explicit connections

Max. 6 connections

 

 

I/O connections

Max. 6 connections (scanner)

 

 

Configuration options for

• DIP switches on the electronics module for IP addressing

measuring device

• Manufacturer-specific software (FieldCare)

 

• Add-on Profile Level 3 for Rockwell Automation control systems

 

• Web browser

 

• Electronic Data Sheet (EDS) integrated in the measuring device

 

 

Configuration of the EtherNet

• Speed: 10 MBit, 100 MBit, auto (factory setting)

interface

• Duplex: half-duplex, full-duplex, auto (factory setting)

 

 

Configuration of the device

• DIP switches on the electronics module for IP addressing (last octet)

address

• DHCP

 

• Manufacturer-specific software (FieldCare)

 

• Add-on Profile Level 3 for Rockwell Automation control systems

 

• Web browser

 

• EtherNet/IP tools, e.g. RSLinx (Rockwell Automation)

 

 

Device Level Ring (DLR)

Yes

 

 

System integration

Information regarding system integration: Operating Instructions

 

→ 135.

 

• Cyclic data transmission

 

• Block model

 

• Input and output groups

 

 

Endress+Hauser

25

 

 

Proline Promass F 500

 

 

 

 

PROFINET

 

 

 

 

 

Protocol

"Application layer protocol for decentral device periphery and distributed

 

 

automation", version 2.3

 

 

 

 

Communication type

100 MBit/s

 

 

 

 

Conformity class

Conformance Class B

 

 

 

 

Netload Class

Netload Class II

 

 

 

 

Baud rates

Automatic 100 Mbit/s with full-duplex detection

 

 

 

 

Cycle times

From 8 ms

 

 

 

 

Polarity

Auto-polarity for automatic correction of crossed TxD and RxD pairs

 

 

 

 

Media Redundancy Protocol

Yes

 

(MRP)

 

 

 

 

 

Device profile

Application interface identifier 0xF600

 

 

Generic device

 

 

 

 

Manufacturer ID

0x11

 

 

 

 

Device type ID

0x843B

 

 

 

 

Device description files (GSD,

Information and files under:

 

DTM, DD)

www.endress.com

 

 

On the product page for the device: Documents/Software → Device drivers

 

 

www.profibus.org

 

 

 

 

Supported connections

• 1 x AR (IO Controller AR)

 

 

• 1 x AR (IO-Supervisor Device AR connection allowed)

 

 

• 1 x Input CR (Communication Relation)

 

 

• 1 x Output CR (Communication Relation)

 

 

• 1 x Alarm CR (Communication Relation)

 

 

 

 

Configuration options for

• DIP switches on the electronics module, for device name assignment (last

 

measuring device

part)

 

 

• Manufacturer-specific software (FieldCare, DeviceCare)

 

 

• Web browser

 

 

• Device master file (GSD), can be read out via the integrated Web server of

 

 

the measuring device

 

 

 

 

Configuration of the

• DIP switches on the electronics module, for device name assignment (last

 

device name

part)

 

 

• DCP protocol

 

 

• Process Device Manager (PDM)

 

 

• Integrated Web server

 

 

 

 

Supported functions

• Identification & Maintenance

 

 

Simple device identification via:

 

 

– Control system

 

 

– Nameplate

 

 

• Measured value status

 

 

The process variables are communicated with a measured value status

 

 

• Blinking feature via the onsite display for simple device identification and

 

 

assignment

 

 

• Device operation via operating tools (e.g. FieldCare, DeviceCare, SIMATIC

 

 

PDM)

 

 

 

 

System integration

Information regarding system integration: Operating Instructions → 135.

 

 

• Cyclic data transmission

 

 

• Overview and description of the modules

 

 

• Status coding

 

 

• Startup configuration

 

 

• Factory setting:

 

 

 

 

FOUNDATION Fieldbus

 

 

 

 

 

Manufacturer ID

0x452B48 (hex)

 

 

 

 

Ident number

0x103B (hex)

 

 

 

26

Endress+Hauser

Proline Promass F 500

Device revision

1

 

 

DD revision

Information and files under:

 

www.endress.com

CFF revision

www.fieldbus.org

 

 

 

Interoperability Test Kit (ITK)

Version 6.2.0

 

 

ITK Test Campaign Number

Information:

 

www.endress.com

 

www.fieldbus.org

 

 

Link Master capability (LAS)

Yes

 

 

Choice of "Link Master" and

Yes

"Basic Device"

Factory setting: Basic Device

 

 

Node address

Factory setting: 247 (0xF7)

 

 

Supported functions

The following methods are supported:

 

• Restart

 

• ENP Restart

 

• Diagnostic

 

• Set to OOS

 

• Set to AUTO

 

• Read trend data

 

• Read event logbook

 

 

Virtual Communication Relationships (VCRs)

 

 

Number of VCRs

44

 

 

Number of link objects in VFD

50

 

 

Permanent entries

1

 

 

Client VCRs

0

 

 

Server VCRs

10

 

 

Source VCRs

43

 

 

Sink VCRs

0

 

 

Subscriber VCRs

43

 

 

Publisher VCRs

43

 

 

Device Link Capabilities

 

 

 

Slot time

4

 

 

Min. delay between PDU

8

 

 

Max. response delay

16

 

 

System integration

Information regarding system integration: Operating Instructions → 135.

 

• Cyclic data transmission

 

• Description of the modules

 

• Execution times

 

• Methods

 

 

Modbus RS485

Protocol

Modbus Applications Protocol Specification V1.1

 

 

Response times

• Direct data access: typically 25 to 50 ms

 

• Auto-scan buffer (data range): typically 3 to 5 ms

 

 

Device type

Slave

 

 

Slave address range

1 to 247

 

 

Broadcast address range

0

 

 

Endress+Hauser

27

Proline Promass F 500

 

Function codes

 

 

• 03: Read holding register

 

 

 

 

 

 

 

 

 

 

 

 

• 04: Read input register

 

 

 

 

 

 

 

 

 

 

 

 

• 06: Write single registers

 

 

 

 

 

 

 

 

 

 

 

 

• 08: Diagnostics

 

 

 

 

 

 

 

 

 

 

 

 

 

• 16: Write multiple registers

 

 

 

 

 

 

 

 

 

 

 

 

• 23: Read/write multiple registers

 

 

 

 

 

 

 

 

 

 

 

 

 

 

 

 

 

 

 

 

Broadcast messages

Supported by the following function codes:

 

 

 

 

 

 

 

 

 

 

 

• 06: Write single registers

 

 

 

 

 

 

 

 

 

 

 

 

• 16: Write multiple registers

 

 

 

 

 

 

 

 

 

 

 

 

• 23: Read/write multiple registers

 

 

 

 

 

 

 

 

 

 

 

 

 

 

 

 

 

 

 

 

 

Supported baud rate

• 1 200 BAUD

 

 

 

 

 

 

 

 

 

 

 

 

 

• 2 400 BAUD

 

 

 

 

 

 

 

 

 

 

 

 

 

• 4 800 BAUD

 

 

 

 

 

 

 

 

 

 

 

 

 

• 9 600 BAUD

 

 

 

 

 

 

 

 

 

 

 

 

 

• 19 200 BAUD

 

 

 

 

 

 

 

 

 

 

 

 

 

• 38 400 BAUD

 

 

 

 

 

 

 

 

 

 

 

 

 

• 57 600 BAUD

 

 

 

 

 

 

 

 

 

 

 

 

 

• 115 200 BAUD

 

 

 

 

 

 

 

 

 

 

 

 

 

 

 

 

 

 

 

 

 

 

 

 

Data transfer mode

 

 

• ASCII

 

 

 

 

 

 

 

 

 

 

 

 

 

 

 

• RTU

 

 

 

 

 

 

 

 

 

 

 

 

 

 

 

 

 

 

 

 

 

 

 

 

Data access

 

 

Each device parameter can be accessed via Modbus RS485.

 

 

 

 

 

 

 

For Modbus register information

 

 

 

 

 

 

Compatibility with

 

 

If the device is replaced, the measuring device Promass 500 supports the

 

earlier model

 

 

compatibility of the Modbus registers for the process variables and the

 

 

 

 

 

 

diagnostic information with the previous model Promass 83. It is not

 

 

 

 

 

 

 

necessary to change the engineering parameters in the automation system.

 

 

 

 

 

 

Description of the function scope of compatibility:

 

 

 

 

 

 

 

 

 

Operating Instructions → 135.

 

 

 

 

 

 

System integration

 

 

Information on system integration: Operating Instructions → 135.

 

 

 

 

 

 

 

• Modbus RS485 information

 

 

 

 

 

 

 

 

 

 

 

 

• Function codes

 

 

 

 

 

 

 

 

 

 

 

 

 

• Register information

 

 

 

 

 

 

 

 

 

 

 

 

• Response time

 

 

 

 

 

 

 

 

 

 

 

 

 

• Modbus data map

 

 

 

 

 

 

 

 

 

 

 

 

 

 

 

 

 

 

 

 

 

 

 

Power supply

 

 

 

 

 

 

 

 

 

 

 

 

 

 

 

 

 

 

 

Terminal assignment

Transmitter: supply voltage, input/outputs

 

 

 

 

 

 

 

 

HART

 

 

 

 

 

 

 

 

 

 

 

 

 

 

 

 

 

 

 

 

 

 

 

 

 

 

 

Supply voltage

 

Input/output

Input/output

Input/output

Input/output

 

 

 

 

1

 

2

 

 

3

4

 

 

 

 

 

 

 

 

 

 

 

 

 

 

 

 

 

 

1 (+)

2 (–)

 

26 (+)

 

27 (–)

24 (+)

 

25 (–)

22 (+)

 

23 (–)

20 (+)

 

21 (–)

 

 

 

 

 

 

 

 

 

 

 

 

 

 

 

 

The terminal assignment depends on the specific device version ordered → 15.

 

 

 

 

 

 

 

 

 

 

 

 

 

 

 

 

FOUNDATION Fieldbus

 

 

 

 

 

 

 

 

 

 

 

 

 

 

 

 

 

Supply voltage

 

Input/output

Input/output

Input/output

Input/output

 

 

 

 

1

 

2

 

 

3

4

 

 

 

 

 

 

 

 

 

 

 

 

 

 

 

 

 

1 (+)

2 (–)

 

26 (A)

 

27 (B)

24 (+)

 

25 (–)

22 (+)

 

23 (–)

20 (+)

 

21 (–)

 

 

 

 

 

 

 

 

 

 

 

 

 

 

 

 

The terminal assignment depends on the specific device version ordered → 15.

 

 

 

 

 

 

 

 

 

 

 

 

 

 

 

 

28

Endress+Hauser

Proline Promass F 500

PROFIBUS PA

Supply voltage

Input/output

Input/output

Input/output

Input/output

 

 

 

1

 

2

 

3

 

4

 

 

 

 

 

 

 

 

 

 

 

 

 

 

1 (+)

2 (–)

26 (B)

 

27 (A)

24 (+)

 

25 (–)

22 (+)

 

23 (–)

20 (+)

 

21 (–)

 

 

 

 

 

 

 

 

 

 

 

 

The terminal assignment depends on the specific device version ordered → 15.

 

 

 

 

 

 

 

 

 

 

 

 

 

 

PROFIBUS DP

Supply voltage

Input/output

Input/output

Input/output

Input/output

 

 

 

1

 

2

 

3

 

4

 

 

 

 

 

 

 

 

 

 

 

 

 

 

1 (+)

2 (–)

26 (B)

 

27 (A)

24 (+)

 

25 (–)

22 (+)

 

23 (–)

20 (+)

 

21 (–)

 

 

 

 

 

 

 

 

 

 

 

 

The terminal assignment depends on the specific device version ordered → 15.

 

 

 

 

 

 

 

 

 

 

 

 

 

 

Modbus RS485

Supply voltage

Input/output

Input/output

Input/output

Input/output

 

 

 

1

 

2

 

3

 

4

 

 

 

 

 

 

 

 

 

 

 

 

 

 

1 (+)

2 (–)

26 (B)

 

27 (A)

24 (+)

 

25 (–)

22 (+)

 

23 (–)

20 (+)

 

21 (–)

 

 

 

 

 

 

 

 

 

 

 

 

The terminal assignment depends on the specific device version ordered → 15.

 

 

 

 

 

 

 

 

 

 

 

 

 

 

PROFINET

Supply voltage

Input/output

Input/output

Input/output

Input/output

 

 

1

 

2

 

3

 

4

 

 

 

 

 

 

 

 

 

 

 

 

1 (+)

2 (–)

PROFINET

24 (+)

 

25 (–)

22 (+)

 

23 (–)

20 (+)

 

21 (–)

 

 

(RJ45 connector)

 

 

 

 

 

 

 

 

 

 

 

The terminal assignment depends on the specific device version

 

 

 

 

 

 

 

 

 

ordered → 15.

 

 

 

 

 

 

 

 

 

 

 

 

 

 

 

EtherNet/IP

Supply voltage

Input/output

Input/output

Input/output

Input/output

 

 

1

 

2

 

3

 

4

 

 

 

 

 

 

 

 

 

 

 

 

1 (+)

2 (–)

EtherNet/IP

24 (+)

 

25 (–)

22 (+)

 

23 (–)

20 (+)

 

21 (–)

 

 

(RJ45 connector)

 

 

 

 

 

 

 

 

 

 

 

The terminal assignment depends on the specific device version

 

 

 

 

 

 

 

 

 

ordered → 15.

 

 

 

 

 

 

 

 

 

 

 

 

 

 

 

Transmitter and sensor connection housing: connecting cable

The sensor and transmitter, which are mounted in separate locations, are interconnected by a connecting cable. The cable is connected via the sensor connection housing and the transmitter housing.

Terminal assignment and connection of the connecting cable:

Proline 500 – digital→ 33

Proline 500 → 34

Device plugs available

Device plugs may not be used in hazardous areas!

Device plugs for fieldbus systems:

Order code for "Input; output 1"

Option SA "FOUNDATION Fieldbus" → 30

Option GA "PROFIBUS PA" → 30

Option RA "PROFINET" → 30

Option NA "EtherNet/IP" → 30

Device plug for connecting to the service interface:

Order code for "Accessory mounted"

option NB, adapter RJ45 M12 (service interface) → 31

Endress+Hauser

29

Proline Promass F 500

Order code for "Input; output 1", option SA "FOUNDATION Fieldbus"

Order code for

Cable entry/connection → 34

 

"Electrical connection"

2

 

3

 

 

 

 

 

M, 3, 4, 5

7/8" connector

 

 

 

 

 

Order code for "Input; output 1", option GA "PROFIBUS PA"

Order code for

Cable entry/connection → 34

 

"Electrical connection"

2

 

3

 

 

 

 

 

L, N, P, U

Connector M12 × 1

 

 

 

 

 

Order code for "Input; output 1", option RA "PROFINET"

Order code for

Cable entry/connection → 34

"Electrical connection"

2

 

3

 

 

 

L, N, P, U

Connector M12 × 1

 

 

 

R 1) 2), S 1) 2), T 1) 2), V 1) 2)

Connector M12 × 1

Connector M12 × 1

1)Cannot be combined with an external WLAN antenna (order code for "Enclosed accessories", option P8) of an RJ45 M12 adapter for the service interface (order code for "Accessories mounted", option NB) or of the remote display and operating module DKX001.

2)Suitable for integrating the device in a ring topology.

Order code for "Input; output 1", option NA "EtherNet/IP"

Order code for

Cable entry/connection → 34

"Electrical connection"

2

 

3

 

 

 

L, N, P, U

Connector M12 × 1

 

 

 

R 1) 2), S 1) 2), T 1) 2), V 1) 2)

Connector M12 × 1

Connector M12 × 1

1)Cannot be combined with an external WLAN antenna (order code for "Enclosed accessories", option P8) of an RJ45 M12 adapter for the service interface (order code for "Accessories mounted", option NB) or of the remote display and operating module DKX001

2)Suitable for integrating the device in a ring topology.

Order code for "Accessory mounted", option NB "Adapter RJ45 M12 (service interface)"

Order code

Cable entry/coupling → 34

 

 

"Accessory mounted"

Cable entry

 

Cable entry

 

 

2

 

3

 

 

 

 

NB

Plug M12 × 1

 

 

 

 

 

Pin assignment, device plug

FOUNDATION Fieldbus

 

 

 

 

 

 

 

 

 

 

 

 

 

 

 

 

 

 

 

 

 

 

 

 

Pin

 

Assignment

Coding

Plug/socket

 

2

 

 

 

3

 

 

 

 

 

 

 

 

 

 

 

 

 

1

+

Signal +

A

Plug

 

 

 

 

 

 

 

 

 

 

 

 

 

1

 

 

 

4

 

2

-

Signal –

 

 

 

 

 

 

 

 

 

 

 

 

 

 

 

 

 

 

 

 

 

 

 

 

 

 

 

3

 

Grounding

 

 

 

 

 

 

 

 

 

 

 

 

 

 

 

 

 

 

 

 

 

4

 

Not assigned

 

 

 

 

 

 

 

 

 

 

 

 

 

 

30

Endress+Hauser

Loading...