This manual describes the Chiller and provides all information required for its safe operation and to
maintain it in good working order.
It is addressed to laboratory personnel and operators in particular.
Read this manual carefully before installing and running your system and note the safety precautions
in section 2 in particular. Store the manual in the immediate vicinity of the instrument, so that it can be
consulted at any time.
No technical modifications may be made to the instrument without the prior written agreement of
BUCHI. Unauthorized modifications may affect the system safety or result in accidents. Technical data
are subject to change without notice.
NOTE
The symbols pertaining to safety (WARNINGS and ATTENTIONS) are explained in section 2.
1 About this manual
This manual is copyright. Information from it may not be reproduced, distributed or used for competitive purposes, nor made available to third parties. The manufacture of any component with the aid of
this manual without prior written agreement is also prohibited.
The English manual is the original language version and serves as basis for all translations into other languages. If you need another language version of this manual, you can
download available versions at www.buchi.com or reorder manuals from a BUCHI representative.
Recirculating Chiller – Types
F-100
(not described in this manual)
F-105
(not described in this manual)
F-108
(not described in this manual)
F-114
(not described in this manual)
F-125
5F-125 Manual, Version B
2 Safety
This section introduces the safety concept of the instrument and contains general rules of behavior
and warnings from direct and indirect hazards concerning the use of the product.
For the users safety, all safety instructions and safety messages in the individual sections shall
be strictly observed and followed. Therefore, the manual must always be available to all persons
performing any tasks described herein.
2.1 User qualification
The instrument may only be used by laboratory personnel and other persons who on account of
training and professional experience know the potential dangers that can develop when operating the
instrument.
Untrained personnel, or persons who are currently being trained, require careful supervision by a qualified person. This Operation Manual serves as a basis for training.
2 Safety
2.2 Proper use
The Recirculating Chiller is conceived and built as a piece of laboratory equipment. Its regulation use is
the cooling of closed cycles (e.g. rotary evaporators, reaction vessels).
When the Recirculating Chiller is used in combination with an instrument (e.g. rotary evaporator and
reaction vessels) all related manuals are to be fully observed.
The regulation use of the Recirculating Chiller also includes its care.
2.3 Improper use
Any other use than the one stated above and any application that does not comply with the technical
data is considered to be improper use. Improper use can cause hazardous situations for the operator
and / or for the instrument and might cause consequential property damage.
The operator bears the sole risk for any damages or hazards caused by improper use!
In particular, the following uses must not be permitted
Installation or use of the instrument in rooms, which require ex-protected instruments.•
The use of spare parts or accessories other than those mentioned in these operating instructions.•
The equipment may not be operated using combustible substances.•
It is not permitted to put anything other than the Rotavapor•
of the Recirculating Chiller.
®
R-151 / 152 / 153 / 220 / 220 SE on top
6F-125 Manual, Version B
2.4 Safety warnings and safety signs used in this manual
!
DANGER, WARNING, CAUTION and NOTICE are standardized signal words for identifying levels of
hazards and risks related to personal injury and property damage. All signal words, which are related
to personal injury are accompanied by the general safety sign.
For your safety it is important to read and fully understand the table below with the different signal
words and their definitions!
Sign Signal wordDefinitionRisk level
DANGER
Indicates a hazardous situation which, if not avoided, will result in
death or serious injury.
2 Safety
★★★★
WARNING
CAUTION
NOTICE
no
Supplementary safety information symbols may be placed in a rectangular panel on the left to the
signal word and the supplementary text (see example below).
Space for
supplementary
safety
information
symbols.
Table of supplementary safety information symbols
The reference list below incorporates all safety information symbols used in this manual and their
meaning.
Indicates a hazardous situation which, if not avoided, could result
in death or serious injury.
Indicates a hazardous situation which, if not avoided, may result
in minor or moderate injury.
Indicates possible property damage, but no
practices related to personal injury.
SIGNAL WORD
Supplementary text, describing the kind and level of hazard / risk seriousness.
List of measures to avoid the described hazard or hazardous situation.•
...•
...•
★★★☆
★★☆☆
★☆☆☆
(property damage only)
SymbolMeaning
General warning
Electrical hazard
Heavy weight, avoid overexertion
Explosive gases, explosive environment
7F-125 Manual, Version B
SymbolMeaning
Fire hazard
Harmful to life forms
Hot item, hot surface
Device damage
2 Safety
Inhalation of substances
Chemical burns by corrosives
Cuts by sharp edges
Flooding
Wear laboratory coat
Wear protective goggles
Wear protective gloves
To be carried by two persons
8F-125 Manual, Version B
Additional user information
!
!
Paragraphs starting with 'NOTE' transport helpful information for working with the device / software or
its supplementaries. NOTEs are not related to any kind of hazard or damage (see following example).
NOTE
Useful tips for the easy operation of the instrument / software.
2.5 Product safety
The Recirculating Chiller has been designed and built in accordance with current state-of-the-art technology, at the time of development. Safety warnings in this manual (as described in section 2.4) serve
to make the user alert to and avoid hazardous situations emanating from residual dangers by giving
appropriate counter measures.
However, risks to users, property and the environment can arise when the instrument is damaged,
used carelessly or improperly.
2.5.1 General hazards
2 Safety
The following safety messages show hazards of general kind which may occur when handling the
instrument. The user shall observe all listed counter measures in order to achieve and maintain the
lowest possible level of hazard.
Additional warning messages can be found whenever actions and situations described in this manual
are related to situational hazards.
DANGER
Death or serious injuries when used in explosive environments.
Do not store or operate the instrument in explosive environments•
Remove all sources of flammable vapors•
Do not store chemicals in the vicinity of the device•
CAUTION
Risk of minor or moderate cuts by sharp edges.
Do not touch defective or broken glassware with bare hands•
Do not touch thin metal edges•
Risk of instrument damage by liquids or mechanical shocks.
9F-125 Manual, Version B
NOTICE
Do not spill liquids over the instrument or its components•
Do not move the instrument when it is loaded with coolant•
Do not drop the instrument or its components•
Keep external vibrations away from the instrument•
Do not operate the instrument without insulation panels and reflector•
Safety messages regarding Refrigerant R134:
!
!
!
!
WARNING
Danger of injury and material damage due to overheating.
Keep at temperature not exceeding 45 °C•
WARNING
Danger of corrosion and poisoning through inhalation of the fumes.
In case of fire and/or explosion do not breathe fumes•
CAUTION
Freeze-burns and eye injuries through direct contact with R134.
Avoid contact with skin and eyes•
Always wear safety goggles•
Always wear safety gloves•
Hoses can be additionally insulated (see optional offering for hose insulation)•
2 Safety
2.5.2 Personal protective equipment
Always wear personal protective equipment such as protective eye goggles and protective clothing.
The personal protective equipment must meet all requirements of all data sheets for the chemicals
used. These instructions are an important part of the Recirculating Chiller and must be made available
at all times to the operating personnel at the place where the equipment is installed.
The instrument is fitted with a thermal overload protection for the compressor.
The minimum quantity of cooling liquid is controlled by a niveau sensor.
10F-125 Manual, Version B
2.6 General safety rules
Responsibility of the operator
The head of the laboratory is responsible for training his/her personnel.
The operator shall inform the manufacturer without delay of any safety-related incidents which might
occur during operation of the instrument or its accessories. Legal regulations, such as local, state and
federal laws applying to the instrument or its accessories must be strictly followed.
Duty of maintenance and care
The operator is responsible for the proper condition of instrument. This includes maintenance, service
and repair jobs that are performed, on schedule by authorized personnel only.
Spare parts to be used
Use only genuine consumables and spare parts for maintenance to assure good system performance,
reliability and safety. Any modifications of spare parts or assemblies are only allowed with the prior
written permission of the manufacturer.
Modifications
Modifications to the instrument are only permitted after prior consultation and with the written approval
of the manufacturer. Modifications and upgrades shall only be carried out by an authorized BUCHI
technical engineer. The manufacturer will decline any claim resulting from unauthorized modifications.
2 Safety
11F-125 Manual, Version B
3 Technical data
This section introduces the reader to the Recirculating Chiller and its specifications. It contains the
technical data, requirements and performance data.
3.1 Technical data
Technical data
Power consumption (max.)2850 W
Supply voltage230 VAC ± 10%
Fuse16 AT
Output socket (max.)2 A
Frequency50/60 Hz
Environmental conditions
- Temperature
- Altitude
- Humidity
Displaydigital, resolution 0.1 °C
Overvoltage categoryII
3 Technical data
For indoor use only
5-40 °C
up to 2000 m
Maximum relative humidity 80% for temperatures up to 31 °C decreasing
linearly to 50% relative humidity at 40 °C
Stainless steel, powder-coated with
polyester-epoxide
PolyesterLaminaPES–
Glass-fiber-reinforced epoxy resinCircuit board–
Polyvinyl chlorideCablePVC–
R134Refrigerant1,1,1,2-tetrafluoro-
990 x 580 x 650 mm
Housing1.4301
ethane
–
Check MSDS
12F-125 Manual, Version B
4 Description of function
This section explains the basic working principle of the Recirculating Chiller. It also shows how the
instrument is structured and provides a general functional description of its assemblies.
4.1 Functional principle
The BUCHI Recirculating Chiller F-125 is a closed-circuit cooler for use with rotary evaporators. The
F-125 has a control unit and display built in to regulate and indicate the actual and set value of the
cooling temperature.
The Recirculating Chiller is designed for installation of a Rotavapor® R-151 / 152 / 153 / 220 / 220 SE on
top of the instrument.
The environmentally-friendly refrigerant R134 is used as the refrigerating medium.
4 Description of function
Process overview F-125
1
6
7
2
3
1 Circulation pump
2 Heat exchanger container
3 Thermostatic expansion valve
4 Filter dryer
5 Heat exchanger
6 Compressor
7 In-line valve
5
4
13F-125 Manual, Version B
5 Putting into operation
!
!
This section describes how the instrument has to be installed. It also gives instructions for the initial
start-up.
NOTE
Inspect the instrument for damage during unpacking. If necessary, prepare a status report immediately and inform customer support and your local BUCHI representative. Keep the original packaging
for future transportation.
Also adhere to all instructions concerning transport as described in section 9.1, Storage and transport.
To move the instrument, loosen the clamps of the wheels, move the instrument and afterwards
secure the wheels by the clamps again.
5.1 Installation site
5 Putting into operation
Put the instrument onto a clean, stable and horizontal surface. Consider the maximum product dimensions and weight. Obtain the environmental conditions as described in section 3.1, technical data.
Installation prerequisites and installation steps:
Do not place any objects in front or behind the instrument.•
The instrument must have a clearance of 40 cm between itself and the wall, both in front and •
behind (sufficient cooling).
Do not place containers, chemicals or other items behind the instrument.•
NOTE
After transport, wait at least 1 hour before switching on the chiller! Within this time the refrigerant •
gathers in the compressor, avoiding damage to the compressor.
To ensure that the power can be cut by unplugging in case of an emergency, the mains plug must •
not be blocked by the instruments or any other items!
Depending on the environmental conditions, condensation water at the cooling tubes and all other •
cold surfaces of the instruments may collect!
DANGER
Death or serious injuries when used in explosive environments.
Do not store or operate the instrument in explosive environments•
Remove all sources of flammable vapors•
Do not store chemicals in the vicinity of the device•
WARNING
Stumbling or falling due to improper installation of the cables and hoses.
The length of the cables and hoses should be kept as short as possible•
If possible, avoid installing cables and hoses in aisle areas•
If installation of cables and hoses in aisle areas is unavoidable, use an adequate protective •
pad in order to avoid the danger of stumbling and damage
14F-125 Manual, Version B
5 Putting into operation
!
!
WARNING
Fire hazard, damage to the instrument through overheating due to inadequate air circulation.
Do not cover the instrument•
Minimum distance from other objects must be at least 40 cm•
CAUTION
Risk of minor or moderate injury due to the heavy weight of the instrument.
Utilize a second person to transport the instrument•
Do not drop the instrument or its transport box•
Place the instrument on a stable, even and vibration-free surface•
Keep limbs out of crushing zone•
5.2 Cooling water routing
Independent of the number of coolers, the cooling water inlet has to be connected in ascending order
starting with the cooler closest to the receiving vessel. The cooling water outlet is always the upper
connector.
Example (simplified schematic)
3
2
1
1 Receiving vessel
2 Inlet (from chiller)
3 Outlet (to chiller)
The drawing shows the cooling water routing for the setup of a single cooler. Other arrangements with
single or double cooler setup can be connected accordingly.
15F-125 Manual, Version B
5.3 Electrical connections
!
The BUCHI Recirculating Chiller F-125 is aimed to be installed stationary and is not equipped with
a power plug. The electrical installation must be performed by an electrician or equivalent specialist.
After the installation procedure an electrical safety test must be performed to verify a safe system
condition such as sufficient grounding.
The used mains circuit has to:
provide the voltage that is given on the type plate of the instrument.•
be able to handle the load of the connected instruments.•
be equipped with adequate fuses and electrical safety measures, in particular proper grounding.•
The Chiller and the Rotavapor•
For details about the installation requirements, see the "Guide for electrical installation" document.
See also technical data (section 3) of all components regarding the different minimum system requirements!
®
should use different phases.
5 Putting into operation
WARNING
Death or serious burns by electric current at installation.
The instrument must be installed by an electrician or equivalent specialist•
After installation, the instrument must be tested for electrical safety•
Risk of instrument damage if mains supply is incorrect.
NOTE
Additional electrical safety measures such as residual current breakers may be necessary to meet •
local laws and regulations!
External power switches (e.g. emergency stop switches) must meet IEC 60947-1 and IEC •
60947-3 requirements. Such devices must be clearly labeled and accessible at any time.
External connections and extension lines must be provided with a grounded conductor lead •
(3-pole couplings, cord or plug equipment). All used power cords used must meet the input
power requirements.
If you have a pump connected to the system the pump can be plugged into the power from the •
Recirculating Chiller F-125.
Notice
External mains supply must meet the voltage given on the type plate•
Check for proper grounding•
Exchange defective cabling immediately•
The max. load on the auxiliary power outlet socket must not exceed 100W!•
16F-125 Manual, Version B
6 Operation
UP
DOWN
START
STOP
This section gives examples of typical instrument applications and instructions on how to operate the
instrument properly and safely. See also section 2.5 “Product safety” for general warnings.
6.1 Operating controls and housing
Front side
6 Operation
5
4
3
2
1
Buttons (not for F-100)
ButtonFunctionality
8
1 Holder for secondary condenser to
6
7
access it from the front side
2 Empty space to install vacuum
pump V-710 (pump not included in
delivery)
3 Cooling media level indicator
4 Operating panel with various
parameters displayed
5 Guidance for Rotavapor
in the right position
6 On-/Off switch (lights green when
instrument is switched on)
7 Cooling lamella for air flow in
8 Wheels (lockable)
®
to install it
Increase the desired set temperature in steps of 0.1 °C.
Decrease the desired set temperature in steps of 0.1 °C.
Button to start or continue the cooling regulation.
Button to stop the cooling regulation.
17F-125 Manual, Version B
Display
6 Operation
3
1
Level indicator
2
2
3
1
4
1 Set temperature of Recirculating Chiller in °C
2 Actual temperature of Recirculating Chiller in °C
3 Active cooling is indicated by the snow flake
4 If the Recirculating Chiller is connected with the
BUCHI vacuum controller the connection symbol
is displayed
1 Lower black line: minimum filling level
2 Upper black line: maximum filling level
3 The floating ball: indicates the actual filling level
of cooling liquid
Rear side
1
2
3
4
L
567
89
J
K
1 Slots for air flow out
2 Main cooling medium flow IN
3 Main cooling medium flow OUT
4 Filling connection for cooling medium
5 RS-485 connection to connect Chiller
®
to the Rotavapor
R-220 SE
6 Resettable fuse
7 Resettable fuse
8 Power cable (to be installed by an
electrician or equivalent specialist)
®
9 Rotavapor
guidances to install
System in the right position.
J Cooling medium flow IN / OUT for
second condenser (bypass it when no
second condenser is installed)
K Output socket i.e. for vacuum pumps
such as BUCHI model V-710
L Wheels (not lockable)
18F-125 Manual, Version B
6.2 Preparing for use
!
!
Prerequisites
All parts must be clean and free of damage.•
Close the drain valve.•
6 Operation
6.2.1 Install the Rotavapor
For installation of the Rotavapor
least two people to lift the Rotavapor® onto the top of the Chiller Trolley. Mount four carry belts on the
Rotavapor® feet . Install one of them on each end. Check for proper seating of the belts. Now lift the
Rotavapor® evenly onto the Chiller Trolley.
6.2.2 Installing the hoses
®
on top of the Chiller
®
on the Chiller you need a transportation specialist. You need at
CAUTION
Risk of minor or moderate injury due to the heavy weight of the instrument.
Utilize a second person to transport the instrument•
Do not drop the instrument or its transport box•
Place the instrument on a stable, even and vibration-free surface•
Keep limbs out of crushing zone•
WARNING
Stumbling or falling due to improper installation of the cables and hoses.
The length of the cables and hoses should be kept as short as possible•
If possible, avoid installing cables and hoses in aisle areas•
If installation of cables and hoses in aisle areas is unavoidable, use an adequate protective •
pad in order to avoid the danger of stumbling and damage
Connect out flow and return flow connections of the Recirculating Chiller with the condenser of the •
rotary evaporator.
NOTE
Only use hoses which can withstand min. -20 °C and 2 bar pressure and secure them with the usual
hose clamps.
®
Setup for a complete Rotavapor
system, includes second condenser on the Chiller and condenser
on the Rotavapor®:
Cooling medium flow IN / OUT for second condenser to the second condenser installed on the •
Chiller.
Connect the main cooling medium flow OUT to the bottom inlet of the cooling coil of the •
®
Rotavapor
Connect the main cooling medium flow IN to the top outlet of the cooling coil of the Rotavapor•
.
®
.
See also section 5.2 “Cooling water routing”.•
19F-125 Manual, Version B
6.2.3 Filling the cooling media tank
!
Do not connect and operate before filling with coolant. Mind the maximum fill level! We recommend a
mixture of ethylene glycol / water at least 30/60, which does not freeze until -16 °C.
NOTE
If you have a large closed cycle cooling system (e.g. long hoses, many condensers connected to the
system in series) fill the refrigerant tank to maximum level.
6 Operation
Risk of instrument damage if wrong cooling medium is used.
NOTICE
Check to ensure that the cooling medium is suitable for use at the desired cooling tempera-•
ture and does not freeze at the working temperature
6.3 Start operation
NOTE
See installation instructions (chapter 5.1) for initial startup!
WARNING
Danger of injury due to tipping or failure of the instrument due to vibrations.
Before starting, the rollers of the instrument must be blocked•
Start the Recirculating Chiller by switching on the Instrument by the main switch. Press start on the
Chiller to fill the cooling coils.
6.4 No BUCHI vacuum controller connected to the system
Preparation steps
System must be in good working order. See final installation check, section 5.1.•
Turn on instrument.•
Check level of cooling transfer medium as described in section 6.1 Level indicator.•
Set the desired working temperature by the UP and DOWN buttons.•
Press the START button to start the cooling process.•
Check the actual temperature on the Chiller, start working when the required temperature is •
reached
20F-125 Manual, Version B
Working steps
The cooling temperature can be adjusted during the operation process by pressing the UP and •
DOWN buttons on the Chiller.
If the Chiller can’t keep the required temperature perform a slower distillation, by means of •
decreasing the vacuum to a slightly higher level so that less vapor is coming to the cooling coil.
Steps for end of process
Press the STOP button on the Chiller.•
The cooling stops immediately. ➡
After a few seconds ventilation also stops. ➡
NOTE
To avoid the rattling of the compressor when the chiller is started, please consider the following:
Keep the chiller connected to the power supply during the entire operation.
6.5 Working with BUCHI vacuum controller
6 Operation
The Chiller model F-125 is equipped with a communication socket. Here, a BUCHI Rotavapor® R-220
SE setup or other appropriate laboratory instruments equipped with a BUCHI vacuum controller
(model V-850 or higher with firmware version v3.0 or higher) can be attached.
To establish communication, switch on all connected devices. After bootup time the UP and DOWN
buttons of the Chillers operation panel are blocked – the Chiller can be controlled via the vacuum
controller buttons.
To Rotavapor /
Vacuum pump
≤ 5m
*click*
NOTE
Do not exceed max. recommended cable length (5 m). For more information about operation see
respective manual of the Vacuum Controller in use.
21F-125 Manual, Version B
7 Maintenance and repairs
!
This section gives instructions on maintenance work to be performed in order to keep the instrument in a good and safe working condition. All maintenance and repair work requiring the opening or
removal of the instrument housing must be carried out by trained service personnel and only with the
tools provided for this purpose.
NOTE
Use only genuine consumables and spare parts for any maintenance and repair work in order to
assure warranty and continued system performance. Any modifications of the Recirculating Chiller or
parts of it require the prior written permission of the manufacturer.
WARNING
Death or serious burns by electric current.
Switch off the instrument, disconnect the power cord and prevent unintentional restart before •
touching any of the elements inside the Recirculating Chiller
Do not spill liquids over the device•
7 Maintenance and repairs
Risk of housing and instrument damage by liquids and detergents.
7.1 Customer service
Only authorized service personnel are allowed to open up the housing and /or perform repair work on
the instrument which is not described in this manual. Authorization requires a comprehensive technical training and knowledge of possible dangers which might arise when working with the instrument.
Such training and knowledge can only be provided by BUCHI.
Addresses of official BUCHI customer service offices are given on the BUCHI website under:
www.buchi.com. If malfunctions occur on your instrument or you have technical questions or application problems, contact one of these offices.
The customer service offers the following:
Spare part delivery•
Repairs•
Technical advice•
NOTICE
Do not spill liquids over the instrument or parts of it•
Wipe off any liquids immediately•
Use only ethanol or soapy water for cleaning•
22
F-125 Manual, Version B
7.2 General inspection and cleaning instructions
!
Check the housing for visible defects (switches, plugs, enclosure etc.) and clean it regularly under
safe conditions with a damp cloth. Wipe off any splashes of aggressive chemicals immediately using
a damp cloth in order to avoid any damage being caused to the coating on the housing. Ethanol as a
cleaning agent is also possible to use.
Cleaning under safe conditions
Switch off the Recirculating Chiller and unplug the power cord.•
Let the system acclimate to ambient temperature completely! ➡
Perform cleaning actions with a damp cloth.•
Regularly clean lamella on front and rear side of the chiller with a damp cloth to remove dust, at •
least once a year.
WARNING
Fire hazard, overheating, or triggering of the over-temperature protection function as a result of a
dirty air intake.
Clean the air intake as needed, but at least 1 x per year.•
7 Maintenance and repairs
Hoses
Check the hoses for wear at least every six months.•
Exchange damaged hoses.•
Cooling media tank
Check prior to use the filling level of the Recirculating Chiller•
Exchange the cooling liquid once a year by means of the drain valve and renew.•
If you use a ethylene glycol / water mixture as cooling media you can drain it to the sink. ➡
23F-125 Manual, Version B
8 Troubleshooting
This section helps to resume operation after a problem has occurred with the instrument which does
not require special technical training. It lists possible occurrences, their probable causes and suggests
how to remedy the problem.
8.1 Error message display
3
1
2
8 Troubleshooting
1 For every error the triangle lights.
2 The error number is indicated on the
display.
3 Error 03 also includes a written message
at the top “COOLING FLUID LOW”.
8.2 Malfunctions and their remedies
The troubleshooting table below lists possible malfunctions and errors of the instrument. The operator
is enabled to correct some of those problems or errors by him/herself. For this, appropriate corrective
measures are listed in the column “Remedy”.
Malfunctions and their remedies
Error CodeProblemRemedy
E01Temperature faultSwitch off the unit, let it cool down, clean air inlet and restart.
E02Temperature faultSwitch off the unit, let it cool down, clean air inlet and restart.
E03No / too little cooling liquid
Pump breakdown
E04Pressure faultSwitch off the unit; let the compressor cool down.
E05Data faultSwitch off the unit and restart.
E06Temperature faultSwitch off the unit, let it cool down, clean air inlet and restart.
Call service if the problem persists.
Call service if the problem persists.
Fill up cooling liquid.
Call service if the problem persists.
Call service to check the system if the problem persists.
Call service if the problem persists.
Call service if the problem persists.
24F-125 Manual, Version B
Malfunctions and their remedies
ProblemRemedy
F-125 does not workSwitch off the unit and let it cool down.
Gently push the resettable fuses back
into prestressed position when fuses
are activated.
Call service if the problem persists.
8 Troubleshooting
1
1 Fuse ok
2 Fuse triggered
2
25F-125 Manual, Version B
9 Shutdown, storage, transport and disposal
!
9 Shutdown, storage, transport and disposal
This section instructs how to shut down and to pack the instrument for storage or transport. Specifications for storage and shipping conditions can also be found listed here.
NOTE
See manufacturer safety data sheet (chapter 11) about the refrigerant!
WARNING
Poisoning or serious injuries through contact with or incorporation of harmful substances.
Wear safety goggles•
Wear safety gloves•
Wear a laboratory coat•
Clean the instrument and all accessories thoroughly to remove possibly dangerous substances•
Do not clean dusty parts with compressed air•
Store the instrument and its accessories in a dry place in the original packaging•
9.1 Storage and transport
Switch off the instrument and unplug the unit. To disassemble the Recirculating Chiller follow the
installation instructions in section 5 in reverse order. Clean the instrument thoroughly! The cooling
liquid is to be drained before storage or shipping. The instrument is to be stored in the original packaging in a dry location. Shipping is also to be done in the original packaging and in upright position
only.
Defective instrument due to improper packaging or improper transport.
Pack the instrument for safe transport with new, suitable packaging material•
In particular, secure the compressor for transport•
Secure the properly packaged instrument on a pallet for transport•
NOTE
After transport, wait at least 1 hour before switching on the chiller! Within this time the refrigerant •
gathers in the compressor, avoiding damage of the compressor.
Drain the cooling medium thoroughly and close the drain cock subsequently.•
NOTICE
26F-125 Manual, Version B
9.2 Disposal
!
Disposal of instrument
For instrument disposal in an environmentally friendly manner, a list of materials is given in section 3.2.
This helps to ensure that the components can be separated and recycled correctly by a specialist for
disposal.
You have to follow valid regional and local laws concerning disposal. For help, please contact your
local authorities!
NOTE
When returning the instrument to the manufacturer for repair work, please copy and complete the
health and safety clearance form on the following page and enclose it with the instrument.
Disposal of Refrigerant R134
The cooling medium R134 must be destroyed in an approved facility, which is equipped to absorb and
neutralize acidic gases and other toxic processing products.
9 Shutdown, storage, transport and disposal
CAUTION
Freeze-burns and eye injuries through direct contact with R134.
Avoid contact with skin and eyes•
Always wear safety goggles•
Always wear safety gloves•
Hoses can be additionally insulated (see optional offering for hose insulation)•
27
F-125 Manual, Version B
9 Shutdown, storage, transport and disposal
Health and Safety Clearance
Health and Safety Clearance_20081110_KESS.doc_20081110 Version 1.0 Page 1/1
Declaration concerning safety, potential hazards and safe disposal of waste.
For the safety and health of our staff, laws and regulations regarding the handling of
dangerous goods, occupational health and safety regulations, safety at work laws and
regulations regarding safe disposal of waste (e.g. chemical waste, chemical residues
or solvents) require that this form must be completed, signed and enclosed to every
return shipment of equipment or defective parts.
Instruments or parts will not be accepted if this declaration is not present.
Equipment
Model: Part/Instrument no.:
1.A Declaration for non dangerous goods
We assure that the returned equipment:
is unused and new.
has not been exposed to toxic, corrosive, biologically active, explosive, radioactive
or other dangerous matters. No hazard emanates from the device!
is free of contamination (e.g.that chemicals, solvents or residues of pumped media
have been drained prior to shipment). No hazard emanates from the device!
1.B Declaration for dangerous goods
Exhaustive list of dangerous substances the equipment has been exposed to:
Chemical, substance Danger classification
We assure that:
all hazardous substances (e.g. toxic, corrosive, biologically active, explosive,
radioactive etc.) which have been processed or been in contact with the equipment
are listed above.
the equipment has been cleaned, decontaminated and is free of transmissible
agents such as hazardous fungi, bacteria, viruses etc. If sterilization is applicable,
all in- and outlets of the equipment have been properly sealed the process.
2. Final Declaration
We hereby declare that:
- we know all about the substances which have been in contact with the equipment
and all questions have been answered correctly.
- we have taken all measures to prevent potential risks that might emanate from the
delivered equipment.
- this document will be attached clearly visible and securely to the outside of the
transport box.
Company name or stamp:
Place, date:
Name (print), job title (print):
Signature:
28F-125 Manual, Version B
10 Spare parts
This section lists spare parts, accessories and options including their ordering information.
Only order spare parts and consumables from BUCHI to maintain the warranty status and to assure
best performance and reliability of the system and affected components. Any modifications to the
spare parts used are only allowed with the prior written permission of the manufacturer.
Always state the product designation, instrument serial and part numbers for warranty clearance when
ordering spare parts!
10.1 Enclosed parts
Enclosed parts list
10 Spare parts
Order no.
mains cables
Control cable
RJ45, 2m
Hose D10/14 3m, 2×004134
Hose nipple
13.5mm, 2×
Screw cap M16x1, 4×019889
Hose clamp, 4×022352
Operation Manual
10.2 Instrument configuration
044989
040329
Single phase, open cable
✓
✓
✓
✓
✓
✓
Content
Instrument version:
BUCHI Recirculating Chiller
Order no.
F-125 Model 2500 controlled 11056478
29F-125 Manual, Version B
10.3 Spare parts, optional accessories
Distil lation Ch il ler
F-125
Ope r ati o n Manual
11593399 en
10 Spare parts
Spare parts
DescriptionOrder no.
Operation Manual, multilingual
English11593399
German11593399
French11593399
Italian11593399
Spanish11593399
Chinese11593399
Distribution head with taps037742
Control cable RJ 45, 2 m044989
Control cable RJ 45, 5 m11056240
Silicone hose D10/14, 2m004134
Hose nipples, 9.5mm046792
Hose nipples, 13.5mm040329
30F-125 Manual, Version B
10 Spare parts
Spare parts
Screw cap M16x1 Rf019889
Hose clamp022352
Insulation for hose D6/1511056888
31F-125 Manual, Version B
11 Declarations and requirements
11.1 FCC requirements (for USA and Canada)
English:
This equipment has been tested and found to comply with the limits for a Class A digital device,
pursuant to both Part 15 of the FCC Rules and the radio interference regulations of the Canadian
Department of Communications. These limits are designed to provide reasonable protection against
harmful interference when the equipment is operated in a commercial environment.
This equipment generates, uses and can radiate radio frequency energy and, if not installed and used
in accordance with the instruction manual, may cause harmful interference to radio communications.
Operation of this equipment in a residential area is likely to cause harmful interference in which case
the user will be required to correct the interference at his own expense.
Français:
Cet appareil a été testé et s'est avéré conforme aux limites prévues pour les appareils numériques
de classe A et à la partie 15 des réglementations FCC ainsi qu’à la réglementation des interférences
radio du Canadian Department of Communications. Ces limites sont destinées à fournir une protection adéquate contre les interférences néfastes lorsque l’appareil est utilisé dans un environnement
commercial.
Cet appareil génère, utilise et peut irradier une énergie à fréquence radioélectrique, il est en outre
susceptible d’engendrer des interférences avec les communications radio, s’il n’est pas installé et
utilisé conformément aux instructions du mode d’emploi. L’utilisation de cet appareil dans les zones
résidentielles peut causer des interférences néfastes, auquel cas l’exploitant sera amené à prendre les
dispositions utiles pour palier aux interférences à ses propres frais.
11 Declarations and requirements
32
F-125 Manual, Version B
11.2 Declaration of conformity
11 Declarations and requirements
33F-125 Manual, Version B
BÜCHI Labortechnik AG
Meierseggstrasse 40
CH-9230 Flawil 1 / Switzerland
T +41 71 394 63 63
F +41 71 394 65 65
www.buchi.com Quality in your hands
Loading...
+ hidden pages
You need points to download manuals.
1 point = 1 manual.
You can buy points or you can get point for every manual you upload.