A grounded terminal which, as far as the operator is concerned, is grounded via a
grounding system.
Protective Earth (PE)
A terminal which must be connected to ground prior to establishing any other
connections.
The ground terminals are situated inside and outside the device:
• Inner ground terminal: Connects the protectiv earth to the mains supply.
• Outer ground terminal: Connects the device to the plant grounding system.
Communication symbols
SymbolMeaning
Wireless Local Area Network (WLAN)
Communication via a wireless, local network.
LED
Light emitting diode is off.
LED
Light emitting diode is on.
LED
Light emitting diode is flashing.
Symbols for certain types of information
SymbolMeaning
Permitted
Procedures, processes or actions that are permitted.
Preferred
Procedures, processes or actions that are preferred.
Forbidden
Procedures, processes or actions that are forbidden.
Tip
Indicates additional information.
Reference to documentation.
Reference to page.
Reference to graphic.
Visual inspection.
4Endress+Hauser
Page 5
Proline Promass F 500
1.
-
.
213
Symbols in graphics
SymbolMeaning
1, 2, 3, ...Item numbers
, 2., 3., …Series of steps
A, B, C, ...Views
A-A, B-B, C-C, ...Sections
Hazardous area
Safe area (non-hazardous area)
Flow direction
Function and system design
Measuring principle
The measuring principle is based on the controlled generation of Coriolis forces. These forces are
always present in a system when both translational and rotational movements are superimposed.
Fc = 2 · ∆m (ν · ω)
Fc = Coriolis force
∆m = moving mass
ω = rotational velocity
ν = radial velocity in rotating or oscillating system
The amplitude of the Coriolis force depends on the moving mass ∆m, its velocity ν in the system and
thus on the mass flow. Instead of a constant rotational velocity ω, the sensor uses oscillation.
In the sensor, two parallel measuring tubes containing flowing fluid oscillate in antiphase, acting like
a tuning fork. The Coriolis forces produced at the measuring tubes cause a phase shift in the tube
oscillations (see illustration):
• At zero flow (when the fluid is at a standstill) the two tubes oscillate in phase (1).
• Mass flow causes deceleration of the oscillation at the inlet of the tubes (2) and acceleration at the
outlet (3).
A0028850
The phase difference (A-B) increases with increasing mass flow. Electrodynamic sensors register the
tube oscillations at the inlet and outlet. System balance is ensured by the antiphase oscillation of the
two measuring tubes. The measuring principle operates independently of temperature, pressure,
viscosity, conductivity and flow profile.
Endress+Hauser5
Page 6
Proline Promass F 500
Density measurement
The measuring tube is continuously excited at its resonance frequency. A change in the mass and
thus the density of the oscillating system (comprising measuring tube and fluid) results in a
corresponding, automatic adjustment in the oscillation frequency. Resonance frequency is thus a
function of medium density. The microprocessor utilizes this relationship to obtain a density signal.
Volume measurement
Together with the measured mass flow, this is used to calculate the volume flow.
Temperature measurement
The temperature of the measuring tube is determined in order to calculate the compensation factor
due to temperature effects. This signal corresponds to the process temperature and is also available
as an output signal.
6Endress+Hauser
Page 7
Proline Promass F 500
2
3
1
A
B
1
2
3
Measuring system
The measuring system consists of a transmitter and a sensor. The transmitter and sensor are
mounted in physically separate locations. They are interconnected by connecting cables.
Transmitter
Two versions of the transmitter are available.
Proline 500 – digitalProline 500
For use in applications not required to meet special requirements due to
ambient or operating conditions.
ANon-hazardous area or Zone 2; Class I, Division 2
BNon-hazardous area or Zone 2; Class I, Division 2 or Zone 1; Class I,
Division 1
1Transmitter
2Connecting cable: cable, separate, standard
3Sensor connection housing with integrated ISEM
• Flexible and cost-effective separate installation.
• A standard cable can be used as the connecting cable.
• Electronics in the transmitter housing, ISEM (intelligent sensor
electronics module) in the sensor connection housing
• Signal transmission: digital
Order code for "Integrated ISEM electronics", option A "Sensor"
Connecting cable (can be ordered in various lengths)→ 132
• Length:
– Zone 2; Class I, Division 2: max. 300 m (1000 ft)
– Zone 1; Class I, Division 1: max. 150 m (500 ft)
• Standard cable with common shield (pair-stranded)
– Process connections: stainless steel, 1.4404 (316/316L); 1.4301
(304); Alloy C22, 2.4602 (UNS N06022)
8Endress+Hauser
Page 9
Proline Promass F 500
2
1
6
5
7
4
3
3
Equipment architecture
A0027512
1Possibilities for integrating measuring devices into a system
1Control system (e.g. PLC)
2Connecting cable (0/4 to 20 mA HART etc.)
3Fieldbus
4Segment coupler
5Non-hazardous area
6Hazardous area: Zone 2; Class I, Division 2
7Hazardous area: Zone 1; Class I, Division 1
SafetyIT security
Our warranty is valid only if the device is installed and used as described in the Operating
Instructions. The device is equipped with security mechanisms to protect it against any inadvertent
changes to the settings.
IT security measures, which provide additional protection for the device and associated data transfer,
must be implemented by the operators themselves in line with their security standards.
Device-specific IT security
The device offers a range of specific functions to support protective measures on the operator's side.
These functions can be configured by the user and guarantee greater in-operation safety if used
correctly. An overview of the most important functions is provided in the following section.
Function/interfaceFactory settingRecommendation
Write protection via hardware write
protection switch → 10
Access code
(also applies for Web server login or
FieldCare connection) → 10
WLAN
(order option in display module)
WLAN security modeEnabled (WPA2-
WLAN passphrase
(password) → 10
WLAN modeAccess PointOn an individual basis following risk
Web server→ 10Enabled.On an individual basis following risk
CDI-RJ45 service interface → 11–On an individual basis following risk
Not enabled.On an individual basis following risk
assessment.
Not enabled
(0000).
Enabled.On an individual basis following risk
PSK)
Serial numberAssign a customized access code during
Assign a customized access code during
commissioning.
assessment.
Do not change.
commissioning.
assessment.
assessment.
assessment.
Endress+Hauser9
Page 10
Proline Promass F 500
Protecting access via hardware write protection
Write access to the device parameters via the local display, Web browser or operating tool (e.g.
FieldCare, DeviceCare) can be disabled via a write protection switch (DIP switch on the
motherboard). When hardware write protection is enabled, only read access to the parameters is
possible.
Hardware write protection is disabled when the device is delivered.
Protecting access via a password
Different passwords are available to protect write access to the device parameters or access to the
device via the WLAN interface.
• User-specific access code
Protect write access to the device parameters via the local display, Web browser or operating tool
(e.g. FieldCare, DeviceCare). Access authorization is clearly regulated through the use of a userspecific access code.
• WLAN passphrase
The network key protects a connection between an operating unit (e.g. notebook or tablet) and the
device via the WLAN interface which can be ordered as an option.
• Infrastructure mode
When the device is operated in infrastructure mode, the WLAN passphrase corresponds to the
WLAN passphrase configured on the operator side.
User-specific access code
Write access to the device parameters via the local display, Web browser or operating tool (e.g.
FieldCare, DeviceCare) can be protected by the modifiable, user-specific access code.
WLAN passphrase: Operation as WLAN access point
A connection between an operating unit (e.g. notebook or tablet) and the device via the WLAN
interface, which can be ordered as an optional extra, is protected by the network key. The WLAN
authentication of the network key complies with the IEEE 802.11 standard.
When the device is delivered, the network key is pre-defined depending on the device. It can be
changed via the WLAN settings submenu in the WLAN passphrase parameter.
Infrastructure mode
A connection between the device and WLAN access point is protected by means of an SSID and
passphrase on the system side. Please contact the relevant system administrator for access.
General notes on the use of passwords
• The access code and network key supplied with the device should be changed during
commissioning.
• Follow the general rules for generating a secure password when defining and managing the access
code or network key.
• The user is responsible for the management and careful handling of the access code and network
key.
Access via Web server
The device can be operated and configured via a Web browser with the integrated Web server. The
connection is via the service interface (CDI-RJ45) or the WLAN interface. For device versions with
the EtherNet/IP and PROFINET communication protocols, the connection can also be established via
the terminal connection for signal transmission with EtherNet/IP or PROFINET (RJ45 connector).
The Web server is enabled when the device is delivered. The Web server can be disabled if necessary
(e.g. after commissioning) via the Web server functionality parameter.
The device and status information can be hidden on the login page. This prevents unauthorized
access to the information.
For detailed information on device parameters, see:
The "Description of Device Parameters" document → 135
Access via OPC-UA
The "OPC UA Server" application package is available in the device version with the HART
communication protocol → 131.
10Endress+Hauser
Page 11
Proline Promass F 500
The device can communicate with OPC UA clients using the "OPC UA Server" application package.
The OPC UA server integrated in the device can be accessed via the WLAN access point using the
WLAN interface - which can be ordered as an optional extra - or the service interface (CDI- RJ45) via
Ethernet network. Access rights and authorization as per separate configuration.
The following Security Modes are supported as per the OPC UA Specification (IEC 62541):
• None
• Basic128Rsa15 – signed
• Basic128Rsa15 – signed and encrypted
Access via service interface (CDI-RJ45)
The device can be connected to a network via the service interface (CDI-RJ45). Device-specific
functions guarantee the secure operation of the device in a network.
The use of relevant industrial standards and guidelines that have been defined by national and
international safety committees, such as IEC/ISA62443 or the IEEE, is recommended. This includes
organizational security measures such as the assignment of access authorization as well as technical
measures such as network segmentation.
The device can be integrated in a ring topology. The device is integrated via the terminal
connection for signal transmission (output 1) and the connection to the service interface (CDIRJ45) → 117.
Endress+Hauser11
Page 12
Input
Measured variableDirect measured variables
• Mass flow
• Density
• Temperature
Calculated measured variables
• Volume flow
• Corrected volume flow
• Reference density
Measuring rangeMeasuring ranges for liquids
Proline Promass F 500
DNMeasuring range full scale values
min(F)
to
[mm][in][kg/h][lb/min]
8³⁄₈0 to 20000 to 73.50
15½0 to 65000 to 238.9
2510 to 180000 to 661.5
401½0 to 450000 to 1654
5020 to 700000 to 2573
8030 to 1800000 to 6615
10040 to 3500000 to 12860
15060 to 8000000 to 29400
250100 to 22000000 to 80850
Measuring range for gases
The full scale value depends on the density and the sound velocity of the gas used and can be
calculated with the formula below:
= minimum (
max(G)
max(G)
max(F)
<
ρ
max(G)
G
max(F)
xConstant dependent on nominal diameter
c
G
d
i
· ρG : x ; ρG · cG · π/2 · (di)2 · 3600)
max(F)
Maximum full scale value for gas [kg/h]
Maximum full scale value for liquid [kg/h]
can never be greater than
max(G)
Gas density in [kg/m³] at operating conditions
Sound velocity (gas) [m/s]
Measuring tube internal diameter [m]
max(F)
max(F)
DNx
[mm][in][kg/m3]
8³⁄₈60
15½80
25190
401½90
50290
803110
12Endress+Hauser
Page 13
Proline Promass F 500
DNx
[mm][in][kg/m3]
1004130
1506200
25010200
To calculate the measuring range, use the Applicator sizing tool → 134
Calculation example for gas
• Sensor: Promass F, DN 50
• Gas: Air with a density of 60.3 kg/m³ (at 20 °C and 50 bar)
• Measuring range (liquid): 70000 kg/h
• x = 90 kg/m³ (for Promass F, DN 50)
Maximum possible full scale value:
Flow rates above the preset full scale value do not override the electronics unit, with the result that
the totalizer values are registered correctly.
Input signalInput and output versions
→ 15
External measured values
To increase the accuracy of certain measured variables or to calculate the corrected volume flow for
gases, the automation system can continuously write different measured values to the measuring
device:
• Operating pressure to increase accuracy (Endress+Hauser recommends the use of a pressure
measuring device for absolute pressure, e.g. Cerabar M or Cerabar S)
• Medium temperature to increase accuracy (e.g. iTEMP)
• Reference density for calculating the corrected volume flow for gases
Various pressure transmitters and temperature measuring devices can be ordered from Endress
+Hauser: see "Accessories" section → 134
It is recommended to read in external measured values to calculate the corrected volume flow.
HART protocol
The measured values are written from the automation system to the measuring device via the HART
protocol. The pressure transmitter must support the following protocol-specific functions:
• HART protocol
• Burst mode
Current input
The measured values are written from the automation system to the measuring device via the
current input → 14.
Digital communication
The measured values can be written from the automation system to the measuring via:
• FOUNDATION Fieldbus
• PROFIBUS DP
• PROFIBUS PA
• Modbus RS485
• EtherNet/IP
• PROFINET
Endress+Hauser13
Page 14
Current input 0/4 to 20 mA
Current input0/4 to 20 mA (active/passive)
Current span• 4 to 20 mA (active)
• 0/4 to 20 mA (passive)
Resolution1 µA
Voltage dropTypically: 0.6 to 2 V for 3.6 to 22 mA (passive)
Maximum input voltage≤ 30 V (passive)
Open-circuit voltage≤ 28.8 V (active)
Possible input variables• Pressure
• Temperature
• Density
Status input
Maximum input values• DC –3 to 30 V
• If status input is active (ON): Ri >3 kΩ
Response timeAdjustable: 5 to 200 ms
Input signal level• Low signal: DC –3 to +5 V
• High signal: DC 12 to 30 V
Assignable functions• Off
• Reset the individual totalizers separately
• Reset all totalizers
• Flow override
Proline Promass F 500
14Endress+Hauser
Page 15
Proline Promass F 500
Output
Output and input variants
Depending on the option selected for output/input 1, different options are available for the other
outputs and inputs. Only one option can be selected for each output/input 1 to 4. The table must be
read vertically (↓).
Example: If the option BA "4–20 mA HART" was selected for output/input 1, one of the options A, B,
D, E, F, H, I or J is available for output 2, and one of the options A, B, D, E, F, H, I or J is available for
output 3 and 4.
Order code for "Output; input 1" (020) →Possible options
Current output 4 to 20 mA HARTBA
Current output 4 to 20 mA HART Ex i↓CA
FOUNDATION Fieldbus↓SA
FOUNDATION Fieldbus Ex i↓TA
PROFIBUS DP↓LA
PROFIBUS PA↓GA
PROFIBUS PA Ex i↓HA
Modbus RS485↓MA
EtherNet/IP 2-port switch integrated↓NA
PROFINET 2-port switch integrated↓RA
Order code for "Output; input 2" (021) →↓↓↓↓↓↓↓↓↓↓
Not assignedAAAAAAAAAA
Current output 0/4 to 20 mABBBBBBB
Current output 0/4 to 20 mA (Ex i)CCC
User configurable input/output
Pulse/frequency/switch outputEEEEEEE
Double pulse output
Pulse/frequency/switch output (Ex i)GGG
Relay outputHHHHHHH
Current input 0/4 to 20 mAIIIIIII
Status inputJJJJJJJ
Order code for "Output; input 3" (022), "Output; input 4" (023)
Not assignedAAAAAAAAAA
Current output 0/4 to 20 mABBBBB
Current output 0/4 to 20 mA (Ex i)C
User configurable input/outputDDDDD
Pulse/frequency/switch outputEEEEE
Double pulse output (slave)
Pulse/frequency/switch output (Ex i)G
Relay outputHHHHH
Current input 0/4 to 20 mAIIIII
Status inputJJJJJ
2)
1)
3)
→↓↓↓↓↓↓↓↓↓↓
4)
DDDDDDD
FF
FF
1)A specific input or output can be assigned to a user configurable input/output → 19.
2)If double pulse output (F) is selected for output/input 2 (021), only the double pulse output (F) option is available for selection for output/input 3
(022).
3)The order code for "Output; input 4" (023) is only available for the Proline 500 – digital transmitter.
4)The double pulse output (F) option is not available for input/output 4.
Endress+Hauser15
Page 16
Output signalHART current output
Current output4 to 20 mA HART
Current spanCan be set to: 4 to 20 mA (active/passive)
Open-circuit voltageDC 28.8 V (active)
Maximum input voltageDC 30 V (passive)
Load250 to 700 Ω
Resolution0.38 µA
DampingConfigurable: 0.07 to 999 s
Assignable measured
variables
Proline Promass F 500
Ex-i, passive
• Mass flow
• Volume flow
• Corrected volume flow
• Density
• Reference density
• Temperature
• Electronic temperature
• Oscillation frequency 0
• Oscillation damping 0
• Signal asymmetry
• Exciter current 0
The range of options increases if the measuring device has one or more
application packages.
PROFIBUS PA
PROFIBUS PAIn accordance with EN 50170 Volume 2, IEC 61158-2 (MBP), galvanically
isolated
Data transmission31.25 kbit/s
Current consumption10 mA
16 mA
Permitted supply voltage9 to 32 V
Bus connectionWith integrated reverse polarity protection
PROFIBUS DP
Signal encodingNRZ code
Data transfer9.6 kBaud…12 MBaud
EtherNet/IP
StandardsIn accordance with IEEE 802.3
PROFINET
StandardsIn accordance with IEEE 802.3
FOUNDATION Fieldbus
FOUNDATION FieldbusH1, IEC 61158-2, galvanically isolated
Data transfer31.25 kbit/s
16Endress+Hauser
Page 17
Proline Promass F 500
Current consumption10 mA
Permitted supply voltage9 to 32 V
Bus connectionWith integrated reverse polarity protection
Modbus RS485
Physical interfaceRS485 in accordance with EIA/TIA-485 standard
Terminating resistorIntegrated, can be activated via DIP switches
Current output 0/4 to 20 mA
Current output0/4 to 20 mA
Maximum output values22.5 mA
Current spanCan be set to:
• 4 to 20 mA (active)
• 0/4 to 20 mA (passive)
Ex-i, passive
Open-circuit voltageDC 28.8 V (active)
Maximum input voltageDC 30 V (passive)
Load0 to 700 Ω
Resolution0.38 µA
DampingAdjustable: 0.07 to 999 s
Assignable measured
variables
• Mass flow
• Volume flow
• Corrected volume flow
• Density
• Reference density
• Temperature
• Electronic temperature
• Oscillation frequency 0
• Oscillation damping 0
• Signal asymmetry
• Exciter current 0
The range of options increases if the measuring device has one or more
application packages.
Pulse/frequency/switch output
FunctionCan be set to pulse, frequency or switch output
VersionOpen collector
Can be set to:
• Active
• Passive
Ex-i, passive
Maximum input valuesDC 30 V, 250 mA (passive)
Open-circuit voltageDC 28.8 V (active)
Voltage dropFor 22.5 mA: ≤ DC 2 V
Pulse output
Maximum input valuesDC 30 V, 250 mA (passive)
Maximum output current22.5 mA (active)
Endress+Hauser17
Page 18
Open-circuit voltageDC 28.8 V (active)
Pulse widthAdjustable: 0.05 to 2000 ms
Maximum pulse rate10000 Impulse/s
Pulse valueAdjustable
Assignable measured
variables
Frequency output
Maximum input valuesDC 30 V, 250 mA (passive)
Maximum output current22.5 mA (active)
Open-circuit voltageDC 28.8 V (active)
Output frequencyAdjustable: end value frequency 2 to 10000 Hz (f
DampingAdjustable: 0 to 999 s
Pulse/pause ratio1:1
Assignable measured
variables
Switch output
Maximum input valuesDC 30 V, 250 mA (passive)
Open-circuit voltageDC 28.8 V (active)
Switching behaviorBinary, conductive or non-conductive
Switching delayAdjustable: 0 to 100 s
Number of switching
cycles
Assignable functions• Off
• Mass flow
• Volume flow
• Corrected volume flow
• Mass flow
• Volume flow
• Corrected volume flow
• Density
• Reference density
• Temperature
• Electronic temperature
• Oscillation frequency 0
• Oscillation damping 0
• Signal asymmetry
• Exciter current 0
The range of options increases if the measuring device has one or more
application packages.
Unlimited
• On
• Diagnostic behavior
• Limit value
– Mass flow
– Volume flow
– Corrected volume flow
– Density
– Reference density
– Temperature
– Totalizer 1-3
• Flow direction monitoring
• Status
– Partially filled pipe detection
– Low flow
The range of options increases if the measuring device has one or more
application packages.
Proline Promass F 500
= 12500 Hz)
max
18Endress+Hauser
Page 19
Proline Promass F 500
Double pulse output
FunctionDouble pulse
VersionOpen collector
Can be set to:
• Active
• Passive
Maximum input valuesDC 30 V, 250 mA (passive)
Open-circuit voltageDC 28.8 V (active)
Voltage dropFor 22.5 mA: ≤ DC 2 V
Output frequencyAdjustable: 0 to 1000 Hz
DampingAdjustable: 0 to 999 s
Pulse/pause ratio1:1
Assignable measured
variables
• Mass flow
• Volume flow
• Corrected volume flow
• Density
• Reference density
• Temperature
The range of options increases if the measuring device has one or more
application packages.
Relay output
FunctionSwitch output
VersionRelay output, galvanically isolated
Switching behaviorCan be set to:
• NO (normally open), factory setting
• NC (normally closed)
Maximum switching
capacity (passive)
Assignable functions• Off
• DC 30 V, 0.1 A
• AC 30 V, 0.5 A
• On
• Diagnostic behavior
• Limit value
– Mass flow
– Volume flow
– Corrected volume flow
– Density
– Reference density
– Temperature
– Totalizer 1-3
• Flow direction monitoring
• Status
– Partially filled pipe detection
– Low flow
The range of options increases if the measuring device has one or more
application packages.
User configurable input/output
One specific input or output is assigned to a user-configurable input/output (configurable I/O)
during device commissioning.
The following inputs and outputs are available for assignment:
• Choice of current output: 4 to 20 mA (active), 0/4 to 20 mA (passive)
• Pulse/frequency/switch output
• Choice of current input: 4 to 20 mA (active), 0/4 to 20 mA (passive)
• Status input
Endress+Hauser19
Page 20
Proline Promass F 500
The technical values correspond to those of the inputs and outputs described in this section.
Signal on alarm
Depending on the interface, failure information is displayed as follows:
HART current output
Device diagnosticsDevice condition can be read out via HART Command 48
PROFIBUS PA
Status and alarm
messages
Error current FDE (Fault
Disconnection Electronic)
Diagnostics in accordance with PROFIBUS PA Profile 3.02
0 mA
PROFIBUS DP
Status and alarm
messages
Diagnostics in accordance with PROFIBUS PA Profile 3.02
EtherNet/IP
Device diagnosticsDevice condition can be read out in Input Assembly
PROFINET
Device diagnosticsAccording to "Application Layer protocol for decentralized periphery", Version 2.3
FOUNDATION Fieldbus
Status and alarm
messages
Error current FDE (Fault
Disconnection Electronic)
Diagnostics in accordance with FF-891
0 mA
Modbus RS485
Failure modeChoose from:
• NaN value instead of current value
• Last valid value
Current output 0/4 to 20 mA
4 to 20 mA
Failure modeChoose from:
• 4 to 20 mA in accordance with NAMUR recommendation NE 43
• 4 to 20 mA in accordance with US
• Min. value: 3.59 mA
• Max. value: 22.5 mA
• Freely definable value between: 3.59 to 22.5 mA
• Actual value
• Last valid value
20Endress+Hauser
Page 21
Proline Promass F 500
0 to 20 mA
Failure modeChoose from:
• Maximum alarm: 22 mA
• Freely definable value between: 0 to 20.5 mA
Pulse/frequency/switch output
Pulse output
Failure modeChoose from:
• Actual value
• No pulses
Frequency output
Failure modeChoose from:
• Actual value
• 0 Hz
• Defined value (f
Switch output
Failure modeChoose from:
• Current status
• Open
• Closed
2 to 12500 Hz)
max
Relay output
Failure modeChoose from:
• Current status
• Open
• Closed
Local display
Plain text displayWith information on cause and remedial measures
BacklightRed backlighting indicates a device error.
Status signal as per NAMUR recommendation NE 107
Interface/protocol
• Via digital communication:
– HART protocol
– FOUNDATION Fieldbus
– PROFIBUS PA
– PROFIBUS DP
– Modbus RS485
– EtherNet/IP
– PROFINET
• Via service interface
– CDI-RJ45 service interface
– WLAN interface
Plain text displayWith information on cause and remedial measures
Additional information on remote operation → 113
Endress+Hauser21
Page 22
Web server
Plain text displayWith information on cause and remedial measures
Light emitting diodes (LED)
Status informationStatus indicated by various light emitting diodes
Ex connection dataSafety-related values
Proline Promass F 500
The following information is displayed depending on the device version:
• Supply voltage active
• Data transmission active
• Device alarm/error has occurred
• EtherNet/IP network available
• EtherNet/IP connection established
• PROFINET network available
• PROFINET connection established
• PROFINET blinking feature
Order code for
Output typeSafety-related values
"Output; input 1"
Option BACurrent output
4 to 20 mA HART
UN = 30 V
UM = 250 V
Option GAPROFIBUS PAUN = 30 V
UM = 250 V
Option LAPROFIBUS DPUN = 30 V
UM = 250 V
Option MAModbus RS485UN = 30 V
UM = 250 V
Option SAFOUNDATION FieldbusUN = 30 V
UM = 250 V
Option NAEtherNet/IPUN = 30 V
UM = 250 V
Option RAPROFINETUN = 30 V
UM = 250 V
Order code for
"Output; input 2";
"Output; input 3"
Output typeSafety-related values
Output; input 2Output; input 3Output; input
"Output; input 4"
24 (+)25 (–)22 (+)23 (–)20 (+)21 (–)
Option BCurrent output
4 to 20 mA
Option DUser configurable input/
output
Option EPulse/frequency/switch
output
Option FDouble pulse outputUN = 30 V
UN = 30 V
UM = 250 V
UN = 30 V
UM = 250 V
UN = 30 V
UM = 250 V
DC
DC
DC
DC
UM = 250 V
Option HRelay outputUN = 30 V
DC
IN =100 mADC/500 mA
UM = 250 V
"Output; input 1"
26 (+)27 (–)
DC
AC
DC
AC
DC
AC
DC
AC
DC
AC
DC
AC
DC
AC
AC
AC
AC
AC
AC
AC
1)
4
22Endress+Hauser
Page 23
Proline Promass F 500
Order code for
"Output; input 2";
"Output; input 3"
Output typeSafety-related values
Output; input 2Output; input 3Output; input
"Output; input 4"
24 (+)25 (–)22 (+)23 (–)20 (+)21 (–)
Option ICurrent input
4 to 20 mA
Option JStatus inputUN = 30 V
UN = 30 V
UM = 250 V
UM = 250 V
DC
AC
DC
AC
1)The order code "Output; input 4" is only available for the Proline 500 – digital transmitter.
Intrinsically safe values
Order code for
"Output; input 1"
Option CACurrent output
Option HAPROFIBUS PA Ex iEx ia
Option TAFOUNDATION Fieldbus
Output typeIntrinsically safe values
"Output; input 1"
26 (+)27 (–)
Ui = 30 V
4 to 20 mA HART Ex i
li = 100 mA
Pi = 1.25 W
Li = 0
Ci = 0
Ex i
1)
Ui = 30 V
li = 570 mA
Pi = 8.5 W
Li = 10 µH
Ci = 5 nF
1)
Ex ia
Ui = 30 V
li = 570 mA
Pi = 8.5 W
Li = 10 µH
Ci = 5 nF
2)
Ex ic
Ui = 32 V
li = 570 mA
Pi = 8.5 W
Li = 10 µH
Ci = 5 nF
2)
Ex ic
Ui = 32 V
li = 570 mA
Pi = 8.5 W
Li = 10 µH
Ci = 5 nF
1)
4
Low flow cut off
Galvanic isolation
1)Only available for the Zone 1; Class I, Division 1 version
2)Only available for the Zone 2; Class I, Division 2 version and only for the Proline 500 – digital transmitter
Order code for
"Output; input 2";
"Output; input 3"
"Output; input 4"
Option CCurrent output
Output typeIntrinsically safe values or NIFW values
Output; input 2Output; input 3 Output; input 4
24 (+)25 (–)22 (+)23 (–)20 (+)21 (–)
Ui = 30 V
4 to 20 mA Ex i
li = 100 mA
Pi = 1.25 W
Li = 0
Ci = 0
Option GPulse/frequency/switch
output Ex i
Ui = 30 V
li = 100 mA
Pi = 1.25 W
Li = 0
Ci = 0
1)The order code "Output; input 4" is only available for the Proline 500 – digital transmitter.
The switch points for low flow cut off are user-selectable.
The outputs are galvanically isolated from one another and from earth (PE).
1)
Endress+Hauser23
Page 24
Protocol-specific dataHART
Manufacturer ID0x11
Device type ID0x3B
HART protocol revision7
Device description files
(DTM, DD)
HART loadMin. 250 Ω
System integrationInformation on system integration: Operating Instructions → 135.
PROFIBUS PA
Manufacturer ID0x11
Ident number0x156D
Profile version3.02
Device description files (GSD,
DTM, DD)
Supported functions• Identification & Maintenance
Configuration of the device
address
Compatibility with
earlier model
System integrationInformation regarding system integration: Operating Instructions → 135.
Proline Promass F 500
Information and files under:
www.endress.com
• Measured variables via HART protocol
• Burst Mode functionality
Information and files under:
• www.endress.com
• www.profibus.org
Simplest device identification on the part of the control system and
nameplate
• PROFIBUS upload/download
Reading and writing parameters is up to ten times faster with PROFIBUS
upload/download
• Condensed status
Simplest and self-explanatory diagnostic information by categorizing
diagnostic messages that occur
• DIP switches on the I/O electronics module
• Local display
• Via operating tools (e.g. FieldCare)
If the device is replaced, the measuring device Promass 500 supports the
compatibility of the cyclic data with previous models. It is not necessary to
adjust the engineering parameters of the PROFIBUS network with the
Promass 500 GSD file.
Earlier models:
• Promass 80 PROFIBUS PA
– ID No.: 1528 (hex)
– Extended GSD file: EH3x1528.gsd
– Standard GSD file: EH3_1528.gsd
• Promass 83 PROFIBUS PA
– ID No.: 152A (hex)
– Extended GSD file: EH3x152A.gsd
– Standard GSD file: EH3_152A.gsd
Description of the function scope of compatibility:
Operating Instructions → 135.
• Cyclic data transmission
• Block model
• Description of the modules
PROFIBUS DP
Manufacturer ID0x11
Ident number0x156F
24Endress+Hauser
Page 25
Proline Promass F 500
Profile version3.02
Device description files (GSD,
DTM, DD)
Supported functions• Identification & Maintenance
Configuration of the device
address
System integrationInformation regarding system integration: Operating Instructions → 135.
Information and files under:
• www.endress.com
On the product page for the device: Documents/Software → Device drivers
• www.profibus.org
Simplest device identification on the part of the control system and
nameplate
• PROFIBUS upload/download
Reading and writing parameters is up to ten times faster with PROFIBUS
upload/download
• Condensed status
Simplest and self-explanatory diagnostic information by categorizing
diagnostic messages that occur
• DIP switches on the I/O electronics module
• Via operating tools (e.g. FieldCare)
• Cyclic data transmission
• Block model
• Description of the modules
EtherNet/IP
Protocol• The CIP Networks Library Volume 1: Common Industrial Protocol
• The CIP Networks Library Volume 2: EtherNet/IP Adaptation of CIP
Communication type• 10Base-T
• 100Base-TX
Device profileGeneric device (product type: 0x2B)
Manufacturer ID0x11
Device type ID0x103B
Baud ratesAutomatic ¹⁰⁄₁₀₀ Mbit with half-duplex and full-duplex detection
PolarityAuto-polarity for automatic correction of crossed TxD and RxD pairs
Supported CIP connectionsMax. 3 connections
Explicit connectionsMax. 6 connections
I/O connectionsMax. 6 connections (scanner)
Configuration options for
measuring device
Configuration of the EtherNet
interface
Configuration of the device
address
Device Level Ring (DLR)Yes
System integrationInformation regarding system integration: Operating Instructions
• DIP switches on the electronics module for IP addressing
• Manufacturer-specific software (FieldCare)
• Add-on Profile Level 3 for Rockwell Automation control systems
• Web browser
• Electronic Data Sheet (EDS) integrated in the measuring device
• Speed: 10 MBit, 100 MBit, auto (factory setting)
• Duplex: half-duplex, full-duplex, auto (factory setting)
• DIP switches on the electronics module for IP addressing (last octet)
• DHCP
• Manufacturer-specific software (FieldCare)
• Add-on Profile Level 3 for Rockwell Automation control systems
• Web browser
• EtherNet/IP tools, e.g. RSLinx (Rockwell Automation)
→ 135.
• Cyclic data transmission
• Block model
• Input and output groups
Endress+Hauser25
Page 26
Proline Promass F 500
PROFINET
Protocol"Application layer protocol for decentral device periphery and distributed
automation", version 2.3
Communication type100 MBit/s
Conformity classConformance Class B
Netload ClassNetload Class II
Baud ratesAutomatic 100 Mbit/s with full-duplex detection
Cycle timesFrom 8 ms
PolarityAuto-polarity for automatic correction of crossed TxD and RxD pairs
Response times• Direct data access: typically 25 to 50 ms
• Auto-scan buffer (data range): typically 3 to 5 ms
Device typeSlave
Slave address range1 to 247
Broadcast address range0
Endress+Hauser27
Page 28
Proline Promass F 500
Function codes• 03: Read holding register
• 04: Read input register
• 06: Write single registers
• 08: Diagnostics
• 16: Write multiple registers
• 23: Read/write multiple registers
Broadcast messagesSupported by the following function codes:
• 06: Write single registers
• 16: Write multiple registers
• 23: Read/write multiple registers
Supported baud rate• 1200 BAUD
• 2400 BAUD
• 4800 BAUD
• 9600 BAUD
• 19200 BAUD
• 38400 BAUD
• 57600 BAUD
• 115200 BAUD
Data transfer mode• ASCII
• RTU
Data accessEach device parameter can be accessed via Modbus RS485.
For Modbus register information
Compatibility with
earlier model
System integrationInformation on system integration: Operating Instructions → 135.
If the device is replaced, the measuring device Promass 500 supports the
compatibility of the Modbus registers for the process variables and the
diagnostic information with the previous model Promass 83. It is not
necessary to change the engineering parameters in the automation system.
Description of the function scope of compatibility:
The terminal assignment depends on the specific device version ordered → 15.
Input/output
2
Input/output
3
Input/output
4
PROFINET
Supply voltageInput/output
1
1 (+)2 (–)PROFINET
(RJ45 connector)
Input/output
2
24 (+)25 (–)22 (+)23 (–)20 (+)21 (–)
The terminal assignment depends on the specific device version
Input/output
3
ordered → 15.
Input/output
4
Device plugs available
EtherNet/IP
Supply voltageInput/output
1
1 (+)2 (–)EtherNet/IP
(RJ45 connector)
Input/output
2
24 (+)25 (–)22 (+)23 (–)20 (+)21 (–)
The terminal assignment depends on the specific device version
Input/output
3
ordered → 15.
Input/output
4
Transmitter and sensor connection housing: connecting cable
The sensor and transmitter, which are mounted in separate locations, are interconnected by a
connecting cable. The cable is connected via the sensor connection housing and the transmitter
housing.
Terminal assignment and connection of the connecting cable:
• Proline 500 – digital→ 33
• Proline 500 → 34
Device plugs may not be used in hazardous areas!
Device plugs for fieldbus systems:
Order code for "Input; output 1"
• Option SA "FOUNDATION Fieldbus" → 30
• Option GA "PROFIBUS PA" → 30
• Option RA "PROFINET" → 30
• Option NA "EtherNet/IP" → 30
Device plug for connecting to the service interface:
Order code for "Accessory mounted"
option NB, adapter RJ45 M12 (service interface) → 31
Endress+Hauser29
Page 30
Order code for "Input; output 1", option SA "FOUNDATION Fieldbus"
1
2
4
3
Order code forCable entry/connection → 34
"Electrical connection"23
M, 3, 4, 57/8" connector–
Order code for "Input; output 1", option GA "PROFIBUS PA"
Order code forCable entry/connection → 34
"Electrical connection"23
L, N, P, UConnector M12 × 1–
Order code for "Input; output 1", option RA "PROFINET"
Order code forCable entry/connection → 34
"Electrical connection"23
L, N, P, UConnector M12 × 1–
1) 2)
1) 2)
1) 2)
R
, S
, T
, V
1) 2)
Connector M12 × 1Connector M12 × 1
Proline Promass F 500
1)Cannot be combined with an external WLAN antenna (order code for "Enclosed accessories", option P8) of
an RJ45 M12 adapter for the service interface (order code for "Accessories mounted", option NB) or of the
remote display and operating module DKX001.
2)Suitable for integrating the device in a ring topology.
Order code for "Input; output 1", option NA "EtherNet/IP"
Order code forCable entry/connection → 34
"Electrical connection"23
L, N, P, UConnector M12 × 1–
1) 2)
1) 2)
1) 2)
R
, S
, T
, V
1) 2)
Connector M12 × 1Connector M12 × 1
1)Cannot be combined with an external WLAN antenna (order code for "Enclosed accessories", option P8) of
an RJ45 M12 adapter for the service interface (order code for "Accessories mounted", option NB) or of the
remote display and operating module DKX001
2)Suitable for integrating the device in a ring topology.
Order code for "Accessory mounted", option NB "Adapter RJ45 M12 (service interface)"
Order codeCable entry/coupling → 34
"Accessory mounted"Cable entry
2
NBPlug M12 × 1–
Cable entry
3
Pin assignment, device plugFOUNDATION Fieldbus
PinAssignmentCodingPlug/socket
1+Signal +APlug
2-Signal –
3Grounding
4Not assigned
30Endress+Hauser
Page 31
Proline Promass F 500
1
2
4
3
3
2
4
3
2
4
3
2
4
PROFIBUS PA
PROFINET
PinAssignmentCodingPlug/socket
1+PROFIBUS PA +APlug
2Grounding
3-PROFIBUS PA –
4Not assigned
PinAssignment
1+TD +
2+RD +
3-TD –
4-RD –
A0032047
CodingPlug/socket
DSocket
Recommended plug:
• Binder, series 763, part no. 99 3729 810 04
• Phoenix, part no. 1543223 SACC-M12MSD-4Q
• When using the device in a hazardous location, use a suitably certified plug.
EtherNet/IP
PinAssignment
1+Tx
2+Rx
3-Tx
4-Rx
A0032047
Recommended plug:
• Binder, series 763, part no. 99 3729 810 04
• Phoenix, part no. 1543223 SACC-M12MSD-4Q
• When using the device in a hazardous location, use a suitably certified plug.
CodingPlug/socket
DSocket
Service interface
Order code for "Accessories mounted", option NB: Adapter RJ45 M12 (service interface)
PinAssignment
1+Tx
2+Rx
3-Tx
4-Rx
A0032047
Recommended plug:
Endress+Hauser31
• Binder, series 763, part no. 99 3729 810 04
• Phoenix, part no. 1543223 SACC-M12MSD-4Q
• When using the device in a hazardous location, use a suitably certified plug.
CodingPlug/socket
DSocket
Page 32
Proline Promass F 500
Supply voltage
Order code for
"Power supply"
Option DDC24 V±20%–
Option EAC100 to 240 V –15…+10%50/60 Hz
Option I
Power consumptionTransmitter
Max. 10 W (active power)
Current consumptionTransmitter
• Max. 400 mA (24 V)
• Max. 200 mA (110 V, 50/60 Hz; 230 V, 50/60 Hz)
Power supply failure
Depending on the device version, the configuration is retained in the device memoryor in the
pluggable data memory (HistoROM DAT).
terminal voltageFrequency range
DC24 V±20%–
AC100 to 240 V –15…+10%50/60 Hz
32Endress+Hauser
Page 33
Proline Promass F 500
+–AB
61 62 63 64
+–AB
61 62 63 64
4
2
1
4
3
5
6
1
2
4
3
5
Electrical connectionConnection of connecting cable: Proline 500 – digital
1Cable entry for cable on transmitter housing
2Protective ground (PE)
3Connecting cable ISEM communication
4Grounding via ground connection; on device plug versions grounding is through the plug itself
5Cable entry for cable or connection of device plug on sensor connection housing
6Protective ground (PE)
A0028198
Depending on the device version of the sensor connection housing, the connecting cable is connected
via terminals or device plugs.
Sensor connection housing
Order code for "Housing"
Option A: aluminum coatedTerminalsTerminals
Option B: stainlessTerminalsTerminals
Option C ultra-compact, hygienic, stainlessDevice plugTerminals
Option L: cast, stainlessTerminalsTerminals
Connection to
sensor connection housing
via
Connection to
transmitter housing
via
Pin assignment, device plug
Device plugs are only available for device version, order code for "Housing":
Option C ultra-compact, hygienic, stainless
For connection to sensor connection housing.
PinColor
1Brown+Supply voltage61
2WhiteA
3BlueB
4Black–Supply voltage63
5––62
1)Cable colors of connecting cable
A connecting cable with a device plug is optionally available.
1)
CodingPlug/socket
APlug
AssignmentConnection
ISEM communication64
to terminal
Endress+Hauser33
Page 34
Proline Promass F 500
4
56
7
8
111241
42
S1S1S2S2GNDTMTMTTTT
++++
++++
S1S1S2S2GNDTMTMTTTT
2
ER
+
+
ER
ER
ER
5
1
10
9
4567891011124142
3
4
X3
X4
X3X4
+
TM
2
TM
2
+
TM
2
TM
2
6
6
12 3456
Connection of the connecting cable: Proline 500
The connecting cable is connected via terminals.
A0033694
1Protective ground (PE)
2Cable entry for connecting cable on transmitter connection housing
3Connecting cable
4Cable entry for connecting cable on sensor connection housing
5Protective ground (PE)
6Terminals X3, X4: temperature sensor; only for device version with order code for "Test, certificate", option JQ
Connecting the transmitter
• Terminal assignment → 28
• Device plug pin assignment → 30
Connecting the Proline 500 – digital transmitter
A0028200
1Terminal connection for supply voltage
2Terminal connection for signal transmission, input/output
3Terminal connection for signal transmission, input/output
4Terminal connection for connecting cable between sensor and transmitter
5Terminal connection for signal transmission, input/output or terminal for network connection (DHCP client)
via service interface (CDI-RJ45); optional: terminal connection for external WLAN antenna
6Protective ground (PE)
An adapter for RJ45 and the M12 plug is optionally available:
Order code for "Accessories", option NB: "Adapter RJ45 M12 (service interface)"
The adapter connects the service interface (CDI-RJ45) to an M12 connector mounted in the
cable entry. Therefore the connection to the service interface can be established via an M12
connector without opening the device.
Network connection (DHCP client) via service interface (CDI-RJ45) → 119
34Endress+Hauser
Page 35
Proline Promass F 500
1
2
3
4
12 3456
Connecting the Proline 500 transmitter
A0026781
1Terminal connection for supply voltage
2Terminal connection for signal transmission, input/output
3Terminal connection for signal transmission, input/output or terminal for network connection (DHCP client)
via service interface (CDI-RJ45); optional: terminal connection for external WLAN antenna
4Protective ground (PE)
An adapter for RJ45 and the M12 plug is optionally available:
Order code for "Accessories", option NB: "Adapter RJ45 M12 (service interface)"
The adapter connects the service interface (CDI-RJ45) to an M12 connector mounted in the
cable entry. Therefore the connection to the service interface can be established via an M12
connector without opening the device.
Network connection (DHCP client) via service interface (CDI-RJ45) → 119
Connecting in a ring topology
Device versions with EtherNet/IP and PROFINET communication protocols can be integrated into a
ring topology. The device is integrated via the terminal connection for signal transmission (output 1)
and the connection to the service interface (CDI-RJ45).
Integrating the transmitter into a ring topology:
• EtherNet/IP → 117
• PROFINET → 118
Transmitter: Proline 500 – digital
A0028200
1Terminal connection for supply voltage
2Terminal connection for signal transmission, input/output
2Terminal connection for signal transmission: PROFINET or EtherNet/IP (RJ45 connector)
4Terminal connection for connecting cable between sensor and transmitter
5Terminal connection to service interface (CDI-RJ45)
6Protective ground (PE)
Endress+Hauser35
Page 36
Proline Promass F 500
1
2
3
4
4
4...20 mA
5
2
1
3
6
2
3
4...20 mA
41
5
Transmitter: Proline 500
1Terminal connection for supply voltage
2Terminal connection for signal transmission: PROFINET or EtherNet/IP (RJ45 connector)
3Terminal connection to service interface (CDI-RJ45)
4Protective ground (PE)
If the device has additional inputs/outputs, these are routed in parallel via the cable entry for
connection to the service interface (CDI-RJ45).
Connection examples
Current output 4 to 20 mA HART
A0026781
A0029055
2Connection example for 4 to 20 mA HART current output (active)
1Automation system with current input (e.g. PLC)
2Cable shield: the cable shield must be grounded at both ends to comply with EMC requirements; observe cable
specifications → 44
3Connection for HART operating devices → 113
4Resistor for HART communication (≥ 250 Ω): observe maximum load → 16
5Analog display unit: observe maximum load → 16
6Transmitter
A0028762
3Connection example for 4 to 20 mA HART current output (passive)
1Automation system with current input (e.g. PLC)
2Power supply
3Cable shield: the cable shield must be grounded at both ends to comply with EMC requirements; observe cable
specifications → 44
36Endress+Hauser
4Analog display unit: observe maximum load → 16
5Transmitter
Page 37
Proline Promass F 500
2
4...20 mA
4
1
2
3
3
6
5
2134
78
66
6
6
5
6
6
5
HART input
A0028763
4Connection example for HART input with a common negative (passive)
1Automation system with HART output (e.g. PLC)
2Active barrier for power supply (e.g. RN221N)
3Cable shield: the cable shield must be grounded at both ends to comply with EMC requirements; observe cable
specifications
4Analog display unit: observe maximum load
5Pressure transmitter (e.g. Cerabar M, Cerabar S): see requirements
6Transmitter
PROFIBUS PA
5Connection example for PROFIBUS PA
1Control system (e.g. PLC)
2PROFIBUS PA segment coupler
3Cable shield: the cable shield must be grounded at both ends to comply with EMC requirements; observe cable
4T-box
5Measuring device
6Local grounding
7Bus terminator
8Potential matching line
specifications
A0028768
Endress+Hauser37
Page 38
Proline Promass F 500
2
1
A
B
3
4
4
A
B
A
B
1
2
4
3
5
5
PROFIBUS DP
A0028765
6Connection example for PROFIBUS DP, non-hazardous area and Zone 2/Div. 2
1Control system (e.g. PLC)
2Cable shield: the cable shield must be grounded at both ends to comply with EMC requirements; observe cable
specifications
3Distribution box
4Transmitter
If baud rates > 1.5 MBaud an EMC cable entry must be used and the cable shield must continue
as far as the terminal wherever possible.
1Control system (e.g. PLC)
2Ethernet switch
3Observe cable specifications → 44
4Connecting cable between the two transmitters
5Transmitter
A0016805
A0027544
Endress+Hauser39
Page 40
FOUNDATION Fieldbus
2134
78
66
6
6
5
6
6
5
2
1
A
B
3
4
4
A
B
A
B
Proline Promass F 500
A0028768
9Connection example for FOUNDATION Fieldbus
1Control system (e.g. PLC)
2Power Conditioner (FOUNDATION Fieldbus)
3Cable shield: the cable shield must be grounded at both ends to comply with EMC requirements; observe cable
10Connection example for Modbus RS485, non-hazardous area and Zone 2; Class I, Division 2
1Control system (e.g. PLC)
2Cable shield: the cable shield must be grounded at both ends to comply with EMC requirements; observe cable
specifications
3Distribution box
4Transmitter
40Endress+Hauser
Page 41
Proline Promass F 500
4...20 mA
2
1
3
2
4...20 mA
3
1
4
1
2
3
12345
Current output 4-20 mA
A0028758
11Connection example for 4-20 mA current output (active)
1Automation system with current input (e.g. PLC)
2Analog display unit: observe maximum load
3Transmitter
12Connection example for 4-20 mA current output (passive)
1Automation system with current input (e.g. PLC)
2Active barrier for power supply (e.g. RN221N)
3Analog display unit: observe maximum load
4Transmitter
Pulse/frequency output
13Connection example for pulse/frequency output (passive)
1Automation system with pulse/frequency input (e.g. PLC)
2Power supply
3Transmitter: Observe input values → 17
A0028759
A0028761
Endress+Hauser41
Page 42
Switch output
1
2
3
1
2
3
4
1
3
2
4
5
14Connection example for switch output (passive)
1Automation system with switch input (e.g. PLC)
2Power supply
3Transmitter: Observe input values → 17
Double pulse output
Proline Promass F 500
A0028760
A0029280
15Connection example for double pulse output (active)
1Automation system with relay input (e.g. PLC)
2Power supply
3Transmitter: Observe input values → 19
Current input
18Connection example for 4 to 20 mA current input
1Power supply
2Terminal box
3External measuring device (for reading in pressure or temperature, for instance)
4Transmitter
Status input
19Connection example for status input
1Automation system with status output (e.g. PLC)
2Power supply
3Transmitter
A0028915
A0028764
Potential equalizationRequirements
No special measures for potential equalization are required.
Endress+Hauser43
Page 44
Please consider the following to ensure correct measurement:
• Same electrical potential for the fluid and sensor
• Company-internal grounding concepts
Proline Promass F 500
terminals
Spring-loaded terminals: Suitable for strands and strands with ferrules.
Conductor cross-section 0.2 to 2.5 mm2 (24 to 12 AWG).
Cable entries
• Cable gland: M20 × 1.5 with cable ⌀ 6 to 12 mm (0.24 to 0.47 in)
• Thread for cable entry:
– NPT ½"
– G ½"
– M20
• Device plug for digital communication: M12
Only available for certain device versions → 29.
• Device plug for connecting cable: M12
A device plug is always used for the device version with the order code for "Sensor connection
housing", option C "Ultra-compact, hygienic, stainless".
Cable specificationPermitted temperature range
• The installation guidelines that apply in the country of installation must be observed.
• The cables must be suitable for the minimum and maximum temperatures to be expected.
Power supply cable
Standard installation cable is sufficient.
Protective ground cable
Cable ≥2.08 mm2 (14 AWG)
The grounding impedance must be less than 1 Ω.
Signal cable
Current output 4 to 20 mA HART
A shielded cable is recommended. Observe grounding concept of the plant.
PROFIBUS PA
Twisted, shielded two-wire cable. Cable type A is recommended .
For further information on planning and installing PROFIBUS networks see:
• Operating Instructions "PROFIBUS DP/PA: Guidelines for planning and commissioning"
(BA00034S)
• PNO Directive 2.092 "PROFIBUS PA User and Installation Guideline"
• IEC 61158-2 (MBP)
PROFIBUS DP
The IEC 61158 standard specifies two types of cable (A and B) for the bus line which can be used for
every transmission rate. Cable type A is recommended.
Cable typeA
Characteristic impedance135 to 165 Ω at a measuring frequency of 3 to 20 MHz
Cable capacitance< 30 pF/m
Wire cross-section
Cable typeTwisted pairs
Loop resistance≤110 Ω/km
> 0.34 mm2 (22 AWG)
44Endress+Hauser
Page 45
Proline Promass F 500
Signal dampingMax. 9 dB over the entire length of the cable cross-section
ShieldCopper braided shielding or braided shielding with foil shield. When grounding
the cable shield, observe the grounding concept of the plant.
For further information on planning and installing PROFIBUS networks see:
• Operating Instructions "PROFIBUS DP/PA: Guidelines for planning and commissioning"
(BA00034S)
• PNO Directive 2.092 "PROFIBUS PA User and Installation Guideline"
• IEC 61158-2 (MBP)
EtherNet/IP
The standard ANSI/TIA/EIA-568-B.2 Annex specifies CAT 5 as the minimum category for a cable
used for EtherNet/IP. CAT 5e and CAT 6 are recommended.
For more information on planning and installing EtherNet/IP networks, please refer to the
"Media Planning and Installation Manual. EtherNet/IP" of ODVA Organization
PROFINET
Standard IEC 61156-6 specifies CAT 5 as the minimum category for a cable used for PROFINET. CAT
5e and CAT 6 are recommended.
For more information on planning and installing PROFINET networks, see: "PROFINET Cabling
and Interconnection Technology", Guideline for PROFINET
FOUNDATION Fieldbus
Twisted, shielded two-wire cable.
For further information on planning and installing FOUNDATION Fieldbus networks see:
• Operating Instructions for "FOUNDATION Fieldbus Overview" (BA00013S)
• FOUNDATION Fieldbus Guideline
• IEC 61158-2 (MBP)
Modbus RS485
The EIA/TIA-485 standard specifies two types of cable (A and B) for the bus line which can be used
for every transmission rate. Cable type A is recommended.
Cable typeA
Characteristic impedance135 to 165 Ω at a measuring frequency of 3 to 20 MHz
Cable capacitance< 30 pF/m
Wire cross-section
Cable typeTwisted pairs
Loop resistance≤110 Ω/km
Signal dampingMax. 9 dB over the entire length of the cable cross-section
ShieldCopper braided shielding or braided shielding with foil shield. When grounding
> 0.34 mm2 (22 AWG)
the cable shield, observe the grounding concept of the plant.
Current output 0/4 to 20 mA
Standard installation cable is sufficient.
Pulse/frequency/switch output
Standard installation cable is sufficient.
Double pulse output
Standard installation cable is sufficient.
Endress+Hauser45
Page 46
Relay output
2
3
C
1
3
B
1
3
A
1
3
A
1
3
B
2
3
C
4
5
6
Standard installation cable is sufficient.
Current input 0/4 to 20 mA
Standard installation cable is sufficient.
Status input
Standard installation cable is sufficient.
Choice of connecting cable between the transmitter and sensor
Depends on the type of transmitter and the installation zones
Proline Promass F 500
1Proline 500 digital transmitter
2Proline 500 transmitter
3Sensor Promass
4Non-hazardous area
5Hazardous area: Zone 2; Class I, Division 2
6Hazardous area: Zone 1; Class I, Division 1
AStandard cable to 500 digital transmitter → 46
Transmitter installed in the non-hazardous area or hazardous area: Zone 2; Class I, Division 2 / sensor
installed in the hazardous area: Zone 2; Class I, Division 2
BStandard cable to 500 digital transmitter → 47
Transmitter installed in the hazardous area: Zone 2; Class I, Division 2 / sensor installed in the hazardous
area: Zone 1; Class I, Division 1
CSignal cable to 500 transmitter → 49
Transmitter and sensor installed in the hazardous area: Zone 2; Class I, Division 2 oder Zone 1;
Class I, Division 1
A: Connecting cable between sensor and transmitter: Proline 500 – digital
Standard cable
A standard cable with the following specifications can be used as the connecting cable.
Design4 cores (2 pairs); uninsulated stranded CU wires; pair-stranded with common
Operating temperatureWhen mounted in a fixed position: –50 to +105 °C (–58 to +221 °F); when cable
can move freely: –25 to +105 °C (–13 to +221 °F)
Available cable lengthFixed: 20 m (65 ft); variable: up to maximum 50 m (165 ft)
1)UV radiation can impair the cable outer sheath. Protect the cable from direct sunshine where possible.
1)
with common shield (2 pairs, pair-
48Endress+Hauser
Page 49
Proline Promass F 500
C: Connecting cable between sensor and transmitter: Proline 500
Standard cable6 × 0.38 mm2 PVC cable
cores
Conductor resistance≤50 Ω/km (0.015 Ω/ft)
Capacitance: core/shield≤420 pF/m (128 pF/ft)
Cable length (max.)20 m (65 ft)
Cable lengths (available for
order)
Operating temperatureDepends on the device version and how the cable is installed:
1)UV radiation can impair the cable outer sheath. Protect the cable from direct sunshine where possible.
5 m (15 ft), 10 m (32 ft), 20 m (65 ft)
• Standard version:
– Cable - fixed installation: –40 to +105 °C (–40 to +221 °F)
– Cable - movable: –25 to +105 °C (–13 to +221 °F)
• Order code for "Test, certificate", option JP:
– Cable - fixed installation: –50 to +105 °C (–58 to +221 °F)
– Cable - movable: –25 to +105 °C (–13 to +221 °F)
• Order code for "Test, certificate", option JQ:
– Cable - fixed installation: –60 to +105 °C (–76 to +221 °F)
– Cable - movable: –25 to +105 °C (–13 to +221 °F)
1)
with common shield and individually shielded
Performance characteristics
Reference operating
conditions
Maximum measured error
• Error limits based on ISO 11631
• Water with +15 to +45 °C (+59 to +113 °F) at2 to 6 bar (29 to 87 psi)
• Specifications as per calibration protocol
• Accuracy based on accredited calibration rigs that are traced to ISO 17025.
To obtain measured errors, use the Applicator sizing tool → 134
o.r. = of reading; 1 g/cm³ = 1 kg/l; T = medium temperature
Base accuracy
Design fundamentals → 53
Mass flow and volume flow (liquids)
±0.05 % o.r. (PremiumCal; order code for "Calibration flow", option D, for mass flow)
±0.10 % o.r.
Mass flow (cryogenic liquids)
Order code for "Measuring tube material", option LA
±0.35 % o.r.
Mass flow (gases)
±0.35 % o.r.
Endress+Hauser49
Page 50
Density (liquids)
Proline Promass F 500
Under
reference operating conditions
[g/cm³][g/cm³][g/cm³]
±0.0005±0.01±0.001
1)Valid over the entire temperature and density range
2)Valid range for special density calibration: 0 to 2 g/cm³, +5 to +80 °C (+41 to +176 °F)
3)Order code for "Application package", option EE "Special density" only in combination with the order code
for "Measuring tube mat., wetted surface", option BB, BF, HA, SA
Standard density
calibration
1)
Wide-range
Density specification
2) 3)
Density (cryogenic liquids)
Order code for "Measuring tube material", option LA
±0.05 g/cm³
Temperature
±0.5 °C ± 0.005 · T °C (±0.9 °F ± 0.003 · (T – 32) °F)
Zero point stability
DNZero point stability
[mm][in][kg/h][lb/min]
8³⁄₈0.0300.001
15½0.2000.007
2510.5400.019
401½2.250.083
5023.500.129
8039.00.330
100414.00.514
150632.01.17
2501088.03.23
High-temperature version: order code for "Measuring tube material", option TT, TU
DNZero point stability
[mm][in][kg/h][lb/min]
2511.800.0661
5027.000.2572
80318.00.6610
For devices with low-temperature version, order code for "Measuring tube mat., wetted surface",
option LA, please note the following:
NOTICE
Zero point confirmation and zero point adjustment are difficult to carry out in the field due to
the vaporization of the cryogenic liquid.
As a general rule, the factory-set zero point should not be changed. Please ensure that the
‣
medium is in the liquid phase if a zero point adjustment is to be carried out.
Flow values
Flow values as turndown parameter depending on nominal diameter.
The outputs have the following base accuracy specifications.
Current output
Accuracy±5 µA
Pulse/frequency output
o.r. = of reading
AccuracyMax. ±50 ppm o.r. (over the entire ambient temperature range)
Repeatability
o.r. = of reading; 1 g/cm3 = 1 kg/l; T = medium temperature
Base repeatability
Design fundamentals → 53
Mass flow and volume flow (liquids)
±0.025 % o.r. (PremiumCal, for mass flow)
±0.05 % o.r.
Endress+Hauser51
Page 52
Mass flow (cryogenic liquids)
Order code for "Measuring tube material", option LA
±0.175 % % o.r.
Mass flow (gases)
±0.25 % o.r.
Density (liquids)
±0.00025 g/cm
3
Density (cryogenic liquids)
Order code for "Measuring tube material", option LA
±0.025 g/cm
3
Temperature
±0.25 °C ± 0.0025 · T °C (±0.45 °F ± 0.0015 · (T–32) °F)
Proline Promass F 500
Response time
Influence of ambient
temperature
Influence of medium
temperature
The response time depends on the configuration (damping).
Current output
Temperature coefficientMax. 1 μA/°C
Pulse/frequency output
Temperature coefficientNo additional effect. Included in accuracy.
Mass flow and volume flow
o.f.s. = of full scale value
When there is a difference between the temperature for zero point adjustment and the process
temperature, the additional measured error of the sensor is typically ±0.0002 % o.f.s./°C (±0.0001 %
o. f.s./°F).
The effect is reduced if zero point adjustment is performed at process temperature.
Density
When there is a difference between the density calibration temperature and the process
temperature, the typical measured error of the sensor is
±0.00005 g/cm3 /°C (±0.000025 g/cm3 /°F). Field density calibration is possible.
Wide-range density specification (special density calibration)
If the process temperature is outside the valid range (→ 49) the measured error is
±0.00005 g/cm3 /°C (±0.000025 g/cm3 /°F)
52Endress+Hauser
Page 53
Proline Promass F 500
3
[°C]
0- 0880 160 240
[° ]F
- 050
100
200
300
0
2
4
6
8
10
2
1
12
14
16
18
320 400 480 560
640
A0027453
1Field density calibration, for example at +20 °C (+68 °F)
2Special density calibration
Temperature
±0.005 · T °C (± 0.005 · (T – 32) °F)
Influence of medium
pressure
The table below shows the effect on accuracy of mass flow due to a difference between calibration
pressure and process pressure.
o.r. = of reading
It is possible to compensate for the effect by:
• Reading in the current pressure measured value via the current input.
• Specifying a fixed value for the pressure in the device parameters.
Operating Instructions→ 135.
DN[% o.r./bar][% o.r./psi]
[mm][in]
83/8no influence
15½no influence
251no influence
401½–0.003–0.0002
502–0.008–0.0006
803–0.009–0.0006
1004–0.007–0.0005
Design fundamentals
1506–0.009–0.0006
25010–0.009–0.0006
o.r. = of reading, o.f.s. = of full scale value
BaseAccu = base accuracy in % o.r., BaseRepeat = base repeatability in % o.r.
MeasValue = measured value; ZeroPoint = zero point stability
Endress+Hauser53
Page 54
Calculation of the maximum measured error as a function of the flow rate
ZeroPoint
BaseAccu
⋅ 100
³
± BaseAccu
ZeroPoint
BaseAccu
⋅ 100
<
ZeroPoint
MeasValue
⋅ 100
±
½ ⋅ ZeroPoint
BaseRepeat
⋅ 100
³
½ ⋅ ZeroPoint
BaseRepeat
⋅ 100
<
ZeroPoint
MeasValue
⋅ 100
½ ⋅±
E [%]
0102030405060
0
0.5
1.0
1.5
2.0
2.5
708090100
Q [%]
Flow rateMaximum measured error in % o.r.
Proline Promass F 500
A0021332
A0021333
Calculation of the maximum repeatability as a function of the flow rate
Flow rateMaximum repeatability in % o.r.
A0021335
A0021336
Example for maximum measured error
A0021339
A0021334
A0021340
A0021337
EMaximum measured error in % o.r. (example with PremiumCal)
QFlow rate in % of maximum full scale value
Installation
No special measures such as supports etc. are necessary. External forces are absorbed by the
construction of the device.
Mounting location
54Endress+Hauser
To prevent measuring errors arising from accumulation of gas bubbles in the measuring tube, avoid
the following mounting locations in the pipe:
• Highest point of a pipeline.
• Directly upstream of a free pipe outlet in a down pipe.
A0028808
A0028772
Page 55
Proline Promass F 500
1
2
3
4
5
Installation in down pipes
However, the following installation suggestion allows for installation in an open vertical pipeline.
Pipe restrictions or the use of an orifice with a smaller cross-section than the nominal diameter
prevent the sensor running empty while measurement is in progress.
A0028773
20Installation in a down pipe (e.g. for batching applications)
1Supply tank
2Sensor
3Orifice plate, pipe restriction
4Valve
5Batching tank
Orientation
DNØ orifice plate, pipe restriction
[mm][in][mm][in]
8³⁄₈60.24
15½100.40
251140.55
401½220.87
502281.10
803501.97
1004652.60
1506903.54
250101505.91
The direction of the arrow on the sensor nameplate helps you to install the sensor according to the
flow direction (direction of medium flow through the piping).
OrientationRecommendation
AVertical orientation
BHorizontal orientation, transmitter at
top
A0015591
1)
Exceptions:
→ 21, 56
A0015589
Endress+Hauser55
Page 56
Proline Promass F 500
12
OrientationRecommendation
CHorizontal orientation, transmitter at
bottom
→ 21, 56
A0015590
DHorizontal orientation, transmitter at
side
A0015592
1)Applications with low process temperatures may decrease the ambient temperature. To maintain the
minimum ambient temperature for the transmitter, this orientation is recommended.
2)Applications with high process temperatures may increase the ambient temperature. To maintain the
maximum ambient temperature for the transmitter, this orientation is recommended.
If a sensor is installed horizontally with a curved measuring tube, match the position of the sensor to
the fluid properties.
2)
Exceptions:
Inlet and outlet runs
A0028774
21Orientation of sensor with curved measuring tube
1Avoid this orientation for fluids with entrained solids: Risk of solids accumulating.
2Avoid this orientation for outgassing fluids: Risk of gas accumulating.
No special precautions need to be taken for fittings which create turbulence, such as valves, elbows
or T-pieces, as long as no cavitation occurs → 69.
56Endress+Hauser
Page 57
Proline Promass F 500
ø 20…70
( 0.79…2.75)ø
~102 (~ 4.0)
4 x
SW 10
3 x
TX 25
149 (5.85)
211 (8.31)
=
5.8 (0.23)
17 (0.67)=
L
5.8 (0.23)
Mounting the transmitter
housing
Proline 500 – digital transmitter
Post mounting
A0029051
22Engineering unit mm (in)
Wall mounting
23Engineering unit mm (in)
LDepends on order code for "Transmitter housing"
Order code for "Transmitter housing"
• Option A, aluminum coated: L =14 mm (0.55 in)
• Option D, polycarbonate: L = 13 mm (0.51 in)
A0029054
Endress+Hauser57
Page 58
Proline Promass F 500
1
4 x
SW 13
!…(! 0.79 to 2.75)20 70
20 (0.79)
100 (3.94)
100 (3.94)
! 8.6 (0.39)
! 18 (0.71)
! 10 (0.39)
Proline 500 transmitter
Post mounting
WARNING
L
Order code for "Transmitter housing", option L "Cast, stainless": cast transmitters are very
heavy.
They are unstable if they are not mounted on a secure, fixed post.
Only mount the transmitter on a secure, fixed post on a stable surface.
‣
Special mounting
instructions
24Engineering unit mm (in)
Wall mounting
25Engineering unit mm (in)
Rupture disk
Information that is relevant to the process: → 69.
A0029057
A0029068
58Endress+Hauser
Page 59
Proline Promass F 500
RUPTURE DISK
i
2
3
1
DN 8 ( ")...150 (6")!⁄#DN 250 (10")
2
3
WARNING
L
Danger from medium escaping!
Medium escaping under pressure can cause injury or material damage.
Take precautions to prevent danger to persons and damage if the rupture disk is actuated.
‣
Observe information on the rupture disk sticker.
‣
Make sure that the function and operation of the rupture disk is not impeded through the
‣
installation of the device.
Do not use a heating jacket.
‣
Do not remove or damage the rupture disk.
‣
The position of the rupture disk is indicated on a sticker beside it.
The transportation guard must be removed.
The existing connecting nozzles are not intended for the purpose of rinsing or pressure monitoring,
but instead serve as the mounting location for the rupture disk.
In the event of a failure of the rupture disk, a discharge device can be screwed onto the internal
thread of the rupture disk in order to drain off any escaping medium.
1Rupture disk label
2Rupture disk with 1/2" NPT internal thread with 1" width across flat
3Transport protection
For information on the dimensions: see the "Mechanical construction" section (accessories)
Zero point adjustment
All measuring devices are calibrated in accordance with state-of-the-art technology. Calibration
takes place under reference conditions→ 49. Therefore, a zero point adjustment in the field is
generally not required.
Experience shows that zero point adjustment is advisable only in special cases:
• To achieve maximum measuring accuracy even with low flow rates.
• Under extreme process or operating conditions (e.g. very high process temperatures or very highviscosity fluids).
A0028903
Endress+Hauser59
Page 60
Environment
Proline Promass F 500
Ambient temperature range
Storage temperature
Climate class
Degree of protection
Measuring device• –40 to +60 °C (–40 to +140 °F)
• Order code for "Test, certificate", option JP:
–50 to +60 °C (–58 to +140 °F)
• Order code for "Test, certificate", option JQ:
– Sensor:–60 to +60 °C (–76 to +140 °F)
– Transmitter: –50 to +60 °C (–58 to +140 °F)
Readability of the
local display
–20 to +60 °C (–4 to +140 °F)
The readability of the display may be impaired at temperatures outside the temperature
range.
Dependency of ambient temperature on medium temperature→ 61
If operating outdoors:
‣
Avoid direct sunlight, particularly in warm climatic regions.
You can order a weather protection cover from Endress+Hauser. → 132.
–50 to +80 °C (–58 to +176 °F)
DIN EN 60068-2-38 (test Z/AD)
Transmitter
• As standard: IP66/67, type 4X enclosure
• When housing is open: IP20, type 1 enclosure
• Display module: IP20, type 1 enclosure
Sensor
• As standard: IP66/67, type 4X enclosure
• With the order code for "Sensor options", option CM: IP69 can also be ordered
External WLAN antenna
IP67
Vibration resistance
Device version: Order code for "Sensor connection housing", option: A "Aluminum, coated", B
"Stainless" or C "Ultra-compact hygienic, stainless"
• Vibration, sinusoidal according to IEC 60068-2-6
– 2 to 8.4 Hz, 7.5 mm peak
– 8.4 to 2000 Hz, 2 g peak
• Vibration broad-band random, according to IEC 60068-2-64
– 10 to 200 Hz, 0.01 g2/Hz
– 200 to 2000 Hz, 0.003 g2/Hz
– Total: 2.70 g rms
Device version: Order code for "Sensor connection housing", option: L "Cast, stainless"
For medium temperature range: standard version
• Vibration, sinusoidal according to IEC 60068-2-6
– 2 to 8.4 Hz, 7.5 mm peak
– 8.4 to 2000 Hz, 2 g peak
• Vibration broad-band random, according to IEC 60068-2-64
– 10 to 200 Hz, 0.01 g2/Hz
– 200 to 2000 Hz, 0.003 g2/Hz
– Total: 2.70 g rms
60Endress+Hauser
Page 61
Proline Promass F 500
Device version: Order code for "Sensor connection housing", option: L "Cast, stainless"
For medium temperature range: extended temperature version, high-temperature version and lowtemperature version
• Vibration, sinusoidal according to IEC 60068-2-6
– 2 to 8.4 Hz, 3.5 mm peak
– 8.4 to 2000 Hz, 1 g peak
• Vibration broad-band random, according to IEC 60068-2-64
– 10 to 200 Hz, 0.003 g2/Hz
– 200 to 2000 Hz, 0.001 g2/Hz
– Total: 1.54 g rms
Shock resistance
Shock resistance
Interior cleaning
Electromagnetic
compatibility (EMC)
Shock, half-sine according to IEC 60068-2-27
6 ms 50 g
Shock due to rough handling following IEC 60068-2-31
• Cleaning in place (CIP)
• Sterilization in place (SIP)
Options
• Oil- and grease-free version for wetted parts, without declaration
Order code for "Service", option HA
• Oil- and grease-free version for wetted parts as per IEC/TR 60877-2.0 and BOC 50000810-4, with
declaration
Order code for "Service", option HB
• As per IEC/EN 61326 and NAMUR Recommendation 21 (NE 21)
• Device version with PROFIBUS DP: Complies with emission limits for industry as per EN 50170
Volume 2, IEC 61784
The following applies for PROFIBUS DP: If baud rates > 1.5 MBaud, an EMC cable entry must be
used and the cable shield must continue as far as the terminal wherever possible.
Details are provided in the Declaration of Conformity.
Process
Medium temperature range
Standard version–50 to +150 °C (–58 to +302 °F)Order code for "Measuring tube
mat., wetted surface", option HA,
SA, SB, SC
Extended temperature version–50 to +240 °C (–58 to +464 °F)Order code for "Measuring tube
mat., wetted surface", option SD, SE,
SF, TH
High-temperature version–50 to +350 °C (–58 to +662 °F)For nominal diameters DN 25 (1"),
DN 50 (2") and DN 80 (3")
Order code for "Measuring tube
mat., wetted surface", option TT, TU
Low-temperature version–196 to +150 °C (–320 to +302 °F)
NOTICE
Material fatigue due to excessive
temperature difference!
Maximum temperature
‣
difference of media used: 300 K
Order code for "Measuring tube
mat., wetted surface", option LA
Endress+Hauser61
Page 62
Proline Promass F 500
T
a
T
m
B
A
Dependency of ambient temperature on medium temperature
26Exemplary representation, values in the table below.
TaAmbient temperature
TmMedium temperature
AMaximum permitted medium temperature Tm at T
require a reduced ambient temperature T
a
BMaximum permitted ambient temperature Ta for the maximum specified medium temperature Tm of the
sensor
= 60 °C (140 °F); higher medium temperatures T
a max
A0031121
m
Values for devices used in the hazardous area:
Separate Ex documentation (XA) for the device → 136.
Not insulatedInsulated
ABAB
1)
Version
Standard version60 °C
Extended temperature version60 °C
High-temperature version60 °C
1)The values apply for Promass F 500 - digital and Promass F 500.
Density
T
a
(140 °F)
T
m
130 °C
(266 °F)
55 °C
(131 °F)
240 °C
(140 °F)
(140 °F)
(464 °F)
240 °C
(464 °F)
50 °C
(122 °F)
0 to 5000 kg/m3 (0 to 312 lb/cf)
T
a
T
m
150 °C
(302 °F)
T
60 °C
(140 °F)
––60 °C
(140 °F)
350 °C
(662 °F)
60 °C
(140 °F)
a
T
m
90 °C
(194 °F)
150 °C
(302 °F)
210 °C
(410 °F)
T
a
45 °C
(113 °F)
50 °C
(122 °F)
50 °C
(122 °F)
T
m
150 °C
(302 °F)
240 °C
(464 °F)
350 °C
(662 °F)
62Endress+Hauser
Page 63
Proline Promass F 500
PN100
PN63
PN40
PN16
0
10
20
30
40
50
60
70
80
90
100
0
200
400
600
800
[bar][psi]
-320
64080160
240
320400480560
[°C]
[°F]
35030025020015010050
0
-200
Pressure-temperature curves
The following pressure/temperature diagrams apply to all pressure-bearing parts of the device and
not just the process connection. The diagrams show the maximum permissible medium pressure
depending on the specific medium temperature.
The medium temperature range of the device depends on the device version → 61.
The values in the pressure-temperature curves only apply for the medium temperature range of
the specific device version.
The extended ambient temperature range for device versions with the order code for "Test,
certificate", option JQ (sensor –60 to +60 °C (–76 to +140 °F)) does not affect the values in the
pressure-temperature curves.
The value for the minimum ambient temperature –60 °C (–76 °F) corresponds to the value for
the minimum medium temperature of the standard version indicated in the particular pressuretemperature curve –50 °C (–58 °F).
• Pressure-temperature curves with temperature range +151 to +240 °C (+304 to +464 °F)
exclusively for extended temperature version of measuring devices.
• Pressure-temperature curves with temperature range +241 to +350 °C (+466 to +662 °F)
exclusively for high temperature version of measuring devices.
• Pressure-temperature curves with temperature range –196 to +150 °C (–320 to +302 °F)
exclusively for low temperature version of measuring devices.
Flange according to EN 1092-1 (DIN 2501)
A0032852-EN
27With flange material 1.4404 (F316/F316L), Alloy C22
Endress+Hauser63
Page 64
Flange according to ASME B16.5
Class 600
Class 300
Class 150
0
10
20
30
40
50
60
70
80
90
100
0
200
400
600
800
[bar][psi]
-320
64080160
240
320400480560
[°C]
[°F]
35030025020015010050
0
-200
Class 600
Class 300
Class 150
-80
640080160
240
320400480560
0
10
20
30
40
50
60
70
80
90
100
0
200
400
600
800
[°C]
[°F]
[bar][psi]
35030025020015010050
0
-50
28With flange material 1.4404 (F316/F316L)
Proline Promass F 500
A0032853-EN
29With flange material Alloy C22
64Endress+Hauser
A0028780-EN
Page 65
Proline Promass F 500
63K
40K
20K
10K
0
10
20
30
40
50
60
70
0
200
400
600
800
[bar][psi]
-320
64080160
240
320400480560
[°C]
[°F]
35030025020015010050
0
-200
DN8…40
≥ DN50
-80
080160
240
320400480
0
10
20
30
40
0
[°C]
[°F]
[bar][psi]
25020015010050
0
-50
50
PN40
-80
080160
240
320400480
0
10
20
30
40
0
[°C]
[°F]
[bar][psi]
25020015010050
0
-50
50
Flange JIS B2220
A0032851-EN
30With flange material 1.4404 (F316/F316L), Alloy C22
Flange DIN 11864-2 Form A
A0028782-EN
31With connection material 1.4404 (316/316L)
Lap joint flange according to EN 1092-1 (DIN 2501)
A0028784-EN
32With flange material 1.4301 (F304); wetted parts Alloy C22
Endress+Hauser65
Page 66
Lap joint flange according to ASME B16.5
Class 150
Class 600
Class 300
-80
080160
240
320400480
0
10
20
30
40
50
60
70
80
90
100
0
200
400
600
800
[°C]
[°F]
[bar][psi]
25020015010050
0
-50
20K
-80
080160
240
320400480
0
10
20
30
0
200
400
[°C]
[°F]
[bar]
[psi]
25020015010050
0
-50
≥ DN50
DN8…40
-80
080160
240
320400480
0
10
20
30
40
0
[°C]
[°F]
[bar][psi]
25020015010050
0
-50
50
33With flange material 1.4301 (F304); wetted parts Alloy C22
Proline Promass F 500
A0028785-EN
Lap joint flange JIS B2220
34With flange material 1.4301 (F304); wetted parts Alloy C22
Thread DIN 11851
66Endress+Hauser
35With connection material 1.4404 (316/316L)
DIN 11851 allows for applications up to +140 °C (+284 °F) if suitable sealing materials are used.
Please take this into account when selecting seals and counterparts, as these components can limit
the pressure and temperature range.
A0028786-EN
A0028794-EN
Page 67
Proline Promass F 500
DN8…40
≥ DN50
-80
080160
240
320400480
0
10
20
30
40
0
[°C]
[°F]
[bar][psi]
25020015010050
0
-50
50
-80
080160
240
320400480
0
10
20
0
[°C]
[°F]
[bar][psi]
25020015010050
0
-50
-80
080160
240
320400480
0
10
20
0
[°C]
[°F]
[bar][psi]
25020015010050
0
-50
Thread DIN 11864-1 Form A
A0028798-EN
36With connection material 1.4404 (316/316L)
Thread ISO 2853
A0028799-EN
37With connection material 1.4404 (316/316L)
Thread SMS 1145
A0028800-EN
38With connection material 1.4404 (316/316L)
SMS 1145 allows for applications up to 16 bar (232 psi) if suitable sealing materials are used. Please
take this into account when selecting seals and counterparts, as these components can limit the
pressure and temperature range.
Endress+Hauser67
Page 68
VCO
60
70
80
90
100
800
[bar][psi]
-80
080160
240
320400480
[°C]
[°F]
25020015010050
0
-50
-80
080160
240
320400480
0
10
20
0
[°C]
[°F]
[bar][psi]
25020015010050
0
-50
39With connection material 1.4404 (316/316L)
Tri-Clamp
Proline Promass F 500
A0028801-EN
The clamp connections are suitable up to a maximum pressure of 16 bar (232 psi). Please observe
the operating limits of the clamp and seal used as they can be over 16 bar (232 psi). The clamp and
seal are not included in the scope of supply.
Sensor housing
For standard versions with the temperature range –50 to +150 °C (–58 to +302 °F), the sensor
housing is filled with dry nitrogen gas and protects the electronics and mechanics inside.
For all other temperature versions the sensor housing is filled with dry inert gas.
If a measuring tube fails (e.g. due to process characteristics like corrosive or abrasive fluids), the
fluid will initially be contained by the sensor housing.
In the event of a tube failure, the pressure level inside the sensor housing will rise according to the
operating process pressure. If the user judges that the sensor housing pressure rating/burst pressure
does not provide an adequate safety margin, the device can be fitted with a rupture disk. This
prevents excessively high pressure from forming inside the sensor housing. Therefore, the use of a
rupture disk is strongly recommended in applications involving high gas pressures, and particularly
in applications in which the process pressure is greater than 2/3 of the sensor housing burst
pressure.
If there is a need to drain the leaking medium into a discharge device, the sensor should be fitted
with a rupture disk. Connect the discharge to the additional threaded connection → 92.
If the sensor is to be purged with gas (gas detection), it should be equipped with purge connections.
Do not open the purge connections unless the containment can be filled immediately with a
dry, inert gas. Use only low pressure to purge. Maximum pressure: 5 bar (72.5 psi).
68Endress+Hauser
Sensor housing nominal pressure rating and burst pressure
The following sensor housing nominal pressure ratings/burst pressures are only valid for standard
devices and/or devices equipped with closed purge connections (not opened/as delivered).
A0032216-EN
Page 69
Proline Promass F 500
If a device fitted with purge connections (order code for "Sensor option", option CH "Purge
connection") is connected to the purge system, the maximum nominal pressure is determined by the
purge system itself or by the device, depending on which component has the lower nominal pressure
classification.
If the device is fitted with a rupture disk (order code for "Sensor option", option CA "Rupture disk"), the
rupture disk trigger pressure is decisive for the maximum nominal pressure .
The sensor housing burst pressure refers to a typical internal pressure which is reached prior to
mechanical failure of the sensor housing and which was determined during type testing. The
corresponding type test declaration can be ordered with the device (order code for "Additional
approval", option LN "Sensor housing burst pressure, type test").
Rupture disk
DNSensor housing nominal
pressure
(designed with a safety factor
≥ 4)
[mm][in][bar][psi][bar][psi]
8³⁄₈405802553698
15½405802002900
251405802804060
401½405801802610
502405801952828
803253621051522
100416232851232
150616232801160
250101014557826
Sensor housing burst pressure
For information on the dimensions: see the "Mechanical construction" section→ 73
To increase the level of safety, a device version with a rupture disk with a trigger pressure of
10 to 15 bar (145 to 217.5 psi)can be used (order code for "Sensor option", option CA "rupture disk").
The use of rupture disks cannot be combined with the separately available heating jacket.
Information on the dimensions of the rupture disk: → 92
Flow limit
Select the nominal diameter by optimizing between the required flow range and permissible pressure
loss.
For an overview of the full scale values for the measuring range, see the "Measuring range"
section → 12
• The minimum recommended full scale value is approx. 1/20 of the maximum full scale value
• In most applications, 20 to 50 % of the maximum full scale value can be considered ideal
• A low full scale value must be selected for abrasive media (such as liquids with entrained solids):
flow velocity < 1 m/s (< 3 ft/s).
• For gas measurement the following rules apply:
– The flow velocity in the measuring tubes should not exceed half the sound velocity (0.5 Mach).
– The maximum mass flow depends on the density of the gas: formula → 12
To calculate the flow limit, use the Applicator sizing tool → 134
Pressure loss
To calculate the pressure loss, use the Applicator sizing tool → 134
Promass F with reduced pressure loss: order code for "Sensor option", option CE "Reduced pressure
loss"
System pressure
It is important that cavitation does not occur, or that gases entrained in the liquids do not outgas.
This is prevented by means of a sufficiently high system pressure.
Endress+Hauser69
Page 70
For this reason, the following mounting locations are recommended:
• At the lowest point in a vertical pipe
• Downstream from pumps (no danger of vacuum)
Proline Promass F 500
A0028777
Thermal insulation
In the case of some fluids, it is important to keep the heat radiated from the sensor to the
transmitter to a low level. A wide range of materials can be used for the required insulation.
The following device versions are recommended for versions with thermal insulation:
• Version with extended neck for insulation:
Order code for "Sensor option", option CG with an extended neck length of 105 mm (4.13 in).
• Extended temperature version:
Order code for "Measuring tube material", option SD, SE, SF or TH with an extended neck length of
105 mm (4.13 in).
• High-temperature version:
Order code for "Measuring tube material", option TT or TU with an extended neck length of
142 mm (5.59 in).
NOTICE
Electronics overheating on account of thermal insulation!
downwards.
Do not insulate the sensor connection housing.
‣
Maximum permissible temperature at the lower end of the sensor connection housing:
‣
80 °C (176 °F)
Thermal insulation with extended neck free: We recommend that you do not insulate the
‣
extended neck in order to ensure optimum dissipation of heat.
A0034391
40Thermal insulation with extended neck free
Low-temperature version: It is generally not necessary to insulate the sensor connection
housing. If insulation is provided, the rules that apply are the same as those for thermal
insulation.
Heating
Some fluids require suitable measures to avoid loss of heat at the sensor.
Heating options
• Electrical heating, e.g. with electric band heaters
• Via pipes carrying hot water or steam
• Via heating jackets
Heating jackets for the sensors can be ordered as accessories from Endress+Hauser .→ 133
70Endress+Hauser
Page 71
Proline Promass F 500
NOTICE
Danger of overheating when heating
Ensure that the temperature at the lower end of the transmitter housing does not exceed
‣
80 °C (176 °F).
Ensure that sufficient convection takes place at the transmitter neck.
‣
Ensure that a sufficiently large area of the transmitted neck remains exposed. The uncovered
‣
part serves as a radiator and protects the electronics from overheating and excessive cooling.
If using in potentially explosive atmospheres, observe the information in the device-specific Ex
‣
documentation. For detailed information on the temperature tables, see the separate document
entitled "Safety Instructions" (XA) for the device.
Vibrations
The high oscillation frequency of the measuring tubes ensures that the correct operation of the
measuring system is not influenced by plant vibrations.
Endress+Hauser71
Page 72
Proline Promass F 500
Custody transfer measurement
The measuring device is optionally tested in accordance with OIML R117/R81 and has an
EU type evaluation certificate which authorizes the use in EU type-examination certificates according
to Measuring Instruments Directive 2014/32/EU for service subject to legal metrological control
("custody transfer") for liquids other than water and cryogenic liquids (Annex VII).
The permitted fluid temperature in these applications is –200 to +90 °C (–328 to +194 °F).
The measuring device is optionally tested in accordance with OIML R137 and has an EU typeexamination certificate according to Measuring Instruments Directive 2014/32/EU for service as a
gas meter subject to legal metrological control ("custody transfer") (Annex IV). The permitted fluid
temperature in these applications is –25 to +55 °C (–13 to +131 °F).
The device is used with a legally controlled totalizer on the local display and optionally with legally
controlled outputs.
Measuring devices subject to legal metrological control totalize in both directions, i.e. all the outputs
consider flow components in the positive (forward) and negative (reverse) flow direction.
Generally a measuring device subject to legal metrological control is secured against tampering by
seals on the transmitter or sensor. These seals may normally only be opened by a representative of
the competent authority for legal metrology controls.
After putting the device into circulation or after sealing the device, operation is only possible to a
limited extent.
Detailed ordering information is available from your local Endress+Hauser sales center for national
approvals, which are based on the OIML certificates, for applications with liquids other than water,
cryogenic liquids or gases.
72Endress+Hauser
Page 73
Proline Promass F 500
N
P
F
G
A
Q
T
G
A
B
C
Q
Q
F
Mechanical construction
Dimensions in SI unitsHousing of Proline 500 – digital transmitter
Non-hazardous area or hazardous area: Zone 2; Class I, Division 2
A0033789
Order code for "Transmitter housing", option A "Aluminum, coated" and order code for "Integrated ISEM
electronics", option A "Sensor"
A
[mm]
167232801872421
F
[mm]
G
[mm]
N
[mm]
P
[mm]
Q
[mm]
Order code for "Transmitter housing", option D "Polycarbonate" and order code for "Integrated ISEM
electronics", option A "Sensor"
A
[mm]
177234901971722
F
[mm]
G
[mm]
N
[mm]
P
[mm]
Q
[mm]
Housing of Proline 500 transmitter
Hazardous area: Zone 2; Class I, Division 2 or Zone 1; Class I, Division 1
Order code for "Transmitter housing", option A "Aluminum, coated" and order code for "Integrated ISEM
A0033788
electronics", option B "Transmitter"
A
[mm]
18885103318217130239
B
[mm]
C
[mm]
F
[mm]
G
[mm]
Q
[mm]
T
[mm]
Endress+Hauser73
Page 74
Proline Promass F 500
A
B
C
G
D
L
E
F
K
M
Order code for "Transmitter housing", option L "Cast, stainless" and order code for "Integrated ISEM
electronics", option B "Transmitter"
A
[mm]
B
[mm]
C
[mm]
F
[mm]
G
[mm]
Q
[mm]
[mm]
18885103295217130239
Sensor with connection housing
Order code for "Sensor connection housing", option A "Aluminum, coated"
DNA
1)
[mm][mm][mm][mm][mm][mm][mm][mm][mm][mm][mm]
81489454751852601365.35
151489454751852601368.30
2514894547518526013612.0
401489454105189.5294.513617.6
501489454141199.5340.513626.0
801489454200219.5419.513640.5
100148945425423849213651.2
150148945437825963713668.9
2501489454548302.5850.5136102.3
1)
B
CDE
2) 3)
F
2) 3)
GKLM
4)
4)
4)
4)
4)
4)
4)
4)
4)
T
A0033787
70
70
70
79
99
139
176
218
305
1)Depending on the cable gland used: values up to + 30 mm
2)With order code for "Sensor option", option CG or order code for "Measuring tube material", option SD, SE,
SF, TH, LA: values +70 mm
3)With order code for "Measuring tube material", option TT, TU: values +104 mm
4)Depends on the process connection in question → 76
Order code for "Sensor connection housing", option B "Stainless"
DNA
[mm][mm][mm][mm][mm][mm][mm][mm][mm][mm][mm]
1)
81377859751802551345.35
151377859751802551348.30
2513778597518025513412.0
401377859105184.5289.513417.6
501377859141194.5335.513426.0
1)
B
CDE
2) 3)
F
2) 3)
GKLM
4)
4)
4)
4)
4)
70
70
70
79
99
74Endress+Hauser
Page 75
Proline Promass F 500
DNA
1)
1)
B
CDE
2) 3)
F
2) 3)
GKLM
[mm][mm][mm][mm][mm][mm][mm][mm][mm][mm][mm]
801377859200214.5414.513440.5
100137785925423348713451.2
150137785937825463213468.9
2501377859548297.5845.5134102.3
4)
4)
4)
4)
139
176
218
305
1)Depending on the cable gland used: values up to + 30 mm
2)With order code for "Sensor option", option CG or order code for "Measuring tube material", option SD, SE,
SF, TH, LA: values +70 mm
3)With order code for "Measuring tube material", option TT, TU: values +104 mm
4)Depends on the process connection in question → 76
Order code for "Sensor connection housing", option C "Ultra-compact hygienic, stainless"
DNA
1)
[mm][mm][mm][mm][mm][mm][mm][mm][mm][mm]
81246856751802551125.35
151246856751802551128.30
2512468567518025511212.0
401246856105184.5289.511217.6
501246856141194.5335.511226.0
801246856200214.5414.511240.5
100124685625423348711251.2
150124685637825463211268.9
2501246856548297.5845.5112102.3
1)
B
CDE
2) 3)
F
2) 3)
GKLM
4)
4)
4)
4)
4)
4)
4)
4)
4)
70
70
70
79
99
139
176
218
305
1)Depending on the cable gland used: values up to + 30 mm
2)With order code for "Sensor option", option CG or order code for "Measuring tube material", option SD, SE,
SF, TH, LA: values +70 mm
3)With order code for "Measuring tube material", option TT, TU: values +104 mm
4)Depends on the process connection in question → 76
Order code for "Sensor connection housing", option L "Cast, stainless"
DNA
1)
[mm][mm][mm][mm][mm][mm][mm][mm][mm][mm][mm]
81458659752082831365.35
151458659752082831368.30
2514586597520828313612.0
401458659105212.5317.513617.6
501458659141222.5363.513626.0
801458659200242.5442.513640.5
100145865925426151513651.2
150145865937828266013668.9
2501458659548325.5873.5136102.3
1)Depending on the cable gland used: values up to + 30 mm
2)With order code for "Sensor option", option CG or order code for "Measuring tube material", option SD, SE,
SF, TH, LA: values +70 mm
3)With order code for "Measuring tube material", option TT, TU: values +104 mm
4)Depends on the process connection in question → 76
1)
B
CDE
2) 3)
F
2) 3)
GKLM
4)
4)
4)
4)
4)
4)
4)
4)
4)
70
70
70
79
99
139
176
218
305
Endress+Hauser75
Page 76
Flange connections
C
D
A
B
E
L
Fixed flange EN 1092-1, ASME B16.5, JIS B2220
Length tolerance for dimension L in mm:
• DN ≤ 100: +1.5 / –2.0
• DN ≥ 125: +3.5
Proline Promass F 500
A0015621
Flange according to EN 1092-1 (DIN 2501): PN16
1.4404 (F316/F316L): order code for "Process connection", option D1S
Alloy C22: order code for "Process connection", option D1C
Flange with groove according to EN 1092-1 Form D (DIN 2512N): PN16
1.4404 (F316/F316L): order code for "Process connection", option D5S
Alloy C22: order code for "Process connection", option D5C
DN
[mm]
1002201808 × ⌀1820107.11127/1400
1502852408 × ⌀2222159.31330/1700
25040535512 × ⌀2626260.41775
Surface roughness (flange): EN 1092-1 Form B1 (DIN 2526 Form C), Ra 3.2 to 12.5 µm
1)Installation length in accordance with NAMUR recommendation NE 132 optionally available (order code
for "Process connection", option D1N or D5N (with groove))
Flange according to EN 1092-1 (DIN 2501): PN16 with reduction in nominal diameter
1.4404 (F316/F316L
DN
[mm]
10080DHS2201808 × ⌀ 1820107.1874
150100DJS2852408 × ⌀ 2222159.31167
200150DLS34029512 × ⌀ 2224206.51461
Surface roughness (flange): EN 1092-1 Form B1 (DIN 2526 Form C), Ra 3.2 to 12.5 µm
A
[mm]
reduction
to DN
[mm]
B
[mm]
Order code for
"Process
connection",
option
C
[mm]
A
[mm]B[mm]
D
[mm]
C
[mm]
E
[mm]
D
[mm]E[mm]L[mm]
L
[mm]
1)
1)
76Endress+Hauser
Page 77
Proline Promass F 500
Flange according to EN 1092-1 (DIN 2501): PN 40
1.4404 (F316/F316L): order code for "Process connection", option D2S
Alloy C22: order code for "Process connection", option D2C
Flange with groove according to EN 1092-1 Form D (DIN 2512N): PN 40
1.4404 (F316/F316L): order code for "Process connection", option D6S
Alloy C22: order code for "Process connection", option D6C
DN
[mm]
1)
8
A
[mm]
B
[mm]
C
[mm]
D
[mm]
E
[mm]
95654 × ⌀141617.3370/510
1595654 × ⌀141617.3404/510
25115854 × ⌀141828.5440/600
401501104 × ⌀181843.1550
501651254 × ⌀182054.5715/715
802001608 × ⌀182482.5840/915
1002351908 × ⌀2224107.11127
1503002508 × ⌀2628159.31370
25045038512 × ⌀3338258.81845
Surface roughness (flange): EN 1092-1 Form B1 (DIN 2526 Form C), Ra 3.2 to 12.5 µm
L
[mm]
2)
2)
2)
2)
2)
1)DN 8 with DN 15 flanges as standard
2)Installation length in accordance with NAMUR recommendation NE 132 optionally available (order code
for "Process connection", option D2N or D6N (with groove))
Flange according to EN 1092-1 (DIN 2501): PN 40 (with DN 25 flanges)
1.4404 (F316/F316L): order code for "Process connection", option R2S
DN
[mm]
A
[mm]
B
[mm]
C
[mm]
D
[mm]
E
[mm]
L
[mm]
8115854 × ⌀141828.5440
15115854 × ⌀141828.5440
Surface roughness (flange): EN 1092-1 Form B1 (DIN 2526 Form C), Ra 3.2 to 12.5 µm
Flange according to EN 1092-1 (DIN 2501): PN 40 with reduction in nominal diameter
1.4404 (F316/F316L)
DN
[mm]
reduction
to DN
[mm]
Order code for
"Process
connection",
A
[mm]B[mm]
C
[mm]
D
[mm]E[mm]L[mm]
option
5040DFS1651254 × ⌀ 182054.5555
8050DGS2001608 × ⌀ 182482.5840
10080DIS2351908 × ⌀ 2224107.1874
150100DKS3002508 × ⌀ 2628159.31167
200150DMS37532012 × ⌀ 3034206.51461
Surface roughness (flange): EN 1092-1 Form B1 (DIN 2526 Form C), Ra 3.2 to 12.5 µm
Endress+Hauser77
Page 78
Proline Promass F 500
Flange according to EN 1092-1 (DIN 2501): PN 63
1.4404 (F316/F316L): order code for "Process connection", option D3S
Alloy C22: order code for "Process connection", option D3C
Flange with groove according to EN 1092-1 Form D (DIN 2512N): PN 63
1.4404 (F316/F316L): order code for "Process connection", option D7S
Alloy C22: order code for "Process connection", option D7C
DN
[mm]
501801354 × ⌀222654.5724
802151708 × ⌀222881.7875
1002502008 × ⌀2630106.31127
1503452808 × ⌀3336157.11410
25047040012 × ⌀3646255.41885
Surface roughness (flange):
EN 1092-1 Form B1 (DIN 2526 Form C), Ra 3.2 to 12.5 µm EN 1092-1 Form B2 (DIN 2526 Form E), Ra
0.8 to 3.2 µm
Flange according to EN 1092-1 (DIN 2501): PN 100
1.4404 (F316/F316L): order code for "Process connection", option D4S
Alloy C22: order code for "Process connection", option D4C
Flange with groove according to EN 1092-1 Form D (DIN 2512N): PN 100
1.4404 (F316/F316L): order code for "Process connection", option D8S
Alloy C22: order code for "Process connection", option D8C
DN
[mm]
1)
8
15105754 × ⌀142017.3420
251401004 × ⌀182428.5470
401701254 × ⌀222642.5590
501951454 × ⌀262853.9740
802301808 × ⌀263280.9885
1002652108 × ⌀3036104.31127
15035529012 × ⌀3344154.01450
Surface roughness (flange): EN 1092-1 Form B2 (DIN 2526 Form E), Ra 0.8 to 3.2 µm
A
[mm]
A
[mm]
105754 × ⌀142017.3400
B
[mm]
B
[mm]
C
[mm]
C
[mm]
D
[mm]
D
[mm]
E
[mm]
E
[mm]
[mm]
[mm]
L
L
1)DN 8 with DN 15 flanges as standard
Flange according to EN 1092-1 (DIN 2501): PN 100
Alloy C22: order code for "Process connection", option D4C
Flange with groove according to EN 1092-1 Form D (DIN 2512N): PN 100
Alloy C22: order code for "Process connection", option D8C
DN
[mm]
25050543012 × ⌀3960248.01949
Surface roughness (flange): EN 1092-1 Form B2 (DIN 2526 Form E), Ra 0.8 to 3.2 µm
A
[mm]
B
[mm]
C
[mm]
D
[mm]
E
[mm]
L
[mm]
78Endress+Hauser
Page 79
Proline Promass F 500
Flange according to ASME B16.5: Class 150
1.4404 (F316/F316L): order code for "Process connection", option AAS
Alloy C22: order code for "Process connection", option AAC
DN
[mm]
1)
8
A
[mm]
B
[mm]
C
[mm]
D
[mm]
E
[mm]
9060.34 × ⌀15.711.215.7370
159060.34 × ⌀15.711.215.7404
2511079.44 × ⌀15.714.226.7440
4012598.44 × ⌀15.717.540.9550
50150120.74 × ⌀19.119.152.6715
80190152.44 × ⌀19.123.978.0840
100230190.58 × ⌀19.123.9102.41127
150280241.38 × ⌀22.425.4154.21398
25040536212 × ⌀25.430.2254.51832
Surface roughness (flange): Ra 3.2 to 6.3 µm
1)DN 8 with DN 15 flanges as standard
L
[mm]
Flange according to ASME B16.5: Class 150 with reduction in nominal diameter
1.4404 (F316/F316L)
DN
[mm]
reduction
to DN
[mm]
Order code for
"Process
connection",
A
[mm]B[mm]
C
[mm]
D
[mm]E[mm]L[mm]
option
5040AHS150120.74 × ⌀ 19.119.152.6550
8050AJS190152.44 × ⌀ 19.123.978.0720
10080ALS230190.58 × ⌀ 19.123.9102.4874
150100ANS280241.38 × ⌀ 22.425.4154.21167
200150APS345298.58 × ⌀ 22.429202.71461
Surface roughness (flange): Ra 3.2 to 6.3 µm
Flange according to ASME B16.5: Class 300
1.4404 (F316/F316L): order code for "Process connection", option ABS
Alloy C22: order code for "Process connection", option ABC
DN
[mm]
1)
8
A
[mm]
B
[mm]
C
[mm]
D
[mm]
E
[mm]
9566.74 × ⌀15.714.215.7370
[mm]
159566.74 × ⌀15.714.215.7404
2512588.94 × ⌀19.117.526.7440
40155114.34 × ⌀22.320.640.9550
501651278 × ⌀19.122.352.6715
80210168.38 × ⌀22.328.478.0840
1002552008 × ⌀22.331.7102.41127
150320269.912 × ⌀22.336.5154.21417
250445387.416 × ⌀28.447.4254.51863
Surface roughness (flange): Ra 3.2 to 6.3 µm
L
1)DN 8 with DN 15 flanges as standard
Endress+Hauser79
Page 80
Proline Promass F 500
Flange according to ASME B16.5: Class 300 with reduction in nominal diameter
1.4404 (F316/F316L)
DN
[mm]
5040AIS1651278 × ⌀ 19.122.352.6615
8050AKS210168.38 × ⌀ 22.328.478.0732
10080AMS2552008 × ⌀ 22.331.7102.4894
150100AOS320269.912 × ⌀ 22.336.5154.21187
200150AQS380330.212 × ⌀ 25.441.7202.71461
Surface roughness (flange): Ra 3.2 to 6.3 µm
Flange according to ASME B16.5: Class 600
1.4404 (F316/F316L): order code for "Process connection", option ACS
Alloy C22: order code for "Process connection", option ACC
Surface roughness (flange): Ra 3.2 to 6.3 µm
reduction
to DN
[mm]
DN
[mm]
1)
8
159566.74 × ⌀15.720.613.9420
2512588.94 × ⌀19.123.924.3490
40155114.34 × ⌀22.328.738.1600
501651278 × ⌀19.131.849.2742
80210168.38 × ⌀22.338.273.7900
100275215.98 × ⌀25.448.497.31157
150355292.112 × ⌀28.447.8154.21467
250510431.816 × ⌀35.169.9254.51946
[mm]
Order code for
"Process
connection",
option
A
9566.74 × ⌀15.720.613.9400
B
[mm]
A
[mm]B[mm]
C
[mm]
C
[mm]
D
[mm]
D
[mm]E[mm]L[mm]
E
[mm]
[mm]
L
1)DN 8 with DN 15 flanges as standard
Flange JIS B2220: 10K
1.4404 (F316/F316L): order code for "Process connection", option NDS
Alloy C22: order code for "Process connection", option NDC
DN
[mm]
501551204 × ⌀191650715
801851508 × ⌀191880832
1002101758 × ⌀19181001127
1502802408 × ⌀23221501354
25040035512 × ⌀25242501775
Surface roughness (flange): Ra 3.2 to 6.3 µm
A
[mm]
B
[mm]
C
[mm]
D
[mm]
E
[mm]
L
[mm]
80Endress+Hauser
Page 81
Proline Promass F 500
Flange JIS B2220: 20K
1.4404 (F316/F316L): order code for "Process connection", option NES
Alloy C22: order code for "Process connection", option NEC
DN
[mm]
1)
8
A
[mm]
B
[mm]
C
[mm]
D
[mm]
E
[mm]
95704 × ⌀151415370
1595704 × ⌀151415404
25125904 × ⌀191625440
401401054 × ⌀191840550
501551208 × ⌀191850715
802001608 × ⌀232280832
1002251858 × ⌀23241001127
15030526012 × ⌀25281501386
25043038012 × ⌀27342501845
Surface roughness (flange): Ra 1.6 to 3.2 µm
1)DN 8 with DN 15 flanges as standard
L
[mm]
Flange JIS B2220: 40K
1.4404 (F316/F316L): order code for "Process connection", option NGS
Alloy C22: order code for "Process connection", option NGC
DN
[mm]
1)
8
A
[mm]
B
[mm]
C
[mm]
D
[mm]
E
[mm]
115804 × ⌀192015400
15115804 × ⌀192015425
25130954 × ⌀192225485
401601204 × ⌀232438600
501651308 × ⌀192650760
802101708 × ⌀233275890
1002502058 × ⌀25361001167
15035529512 × ⌀33441501498
Surface roughness (flange): Ra 1.6 to 3.2 µm
1)DN 8 with DN 15 flanges as standard
Flange JIS B2220: 63K
1.4404 (F316/F316L): order code for "Process connection", option NHS
Alloy C22: order code for "Process connection", option NHC
DN
[mm]
1)
8
A
[mm]
B
[mm]
C
[mm]
D
[mm]
E
[mm]
120854 × ⌀192312420
15120854 × ⌀192312440
251401004 × ⌀232722494
401751304 × ⌀253235620
501851458 × ⌀233448775
802301858 × ⌀254073915
1002702208 × ⌀2744981167
L
[mm]
L
[mm]
Endress+Hauser81
Page 82
Proline Promass F 500
Flange JIS B2220: 63K
1.4404 (F316/F316L): order code for "Process connection", option NHS
Alloy C22: order code for "Process connection", option NHC
DN
[mm]
15036530512 × ⌀33541461528
Surface roughness (flange): Ra 1.6 to 3.2 µm
1)DN 8 with DN 15 flanges as standard
A
[mm]
B
[mm]
C
[mm]
D
[mm]
E
[mm]
L
[mm]
82Endress+Hauser
Page 83
Proline Promass F 500
C
D
A
B
E
XX
L
Fixed flange DIN 11864-2
A0015627
41Detail X: Asymmetrical process connection; the part shown in blue is provided by the supplier.
Length tolerance for dimension L in mm:
+1.5 / –2.0
Flange DIN11864-2 Form A, for pipe according to DIN11866 series A, flat with notch
1.4404 (316/316L)
Order code for "Process connection", option KCS
DN
[mm]
854374 × ⌀91010387
1559424 × ⌀91016418
2570534 × ⌀91026454
4082654 × ⌀91038560
5094774 × ⌀91050720
801331128 × ⌀111281900
1001591378 × ⌀11141001127
3A-version available: order code for "Additional approval", option LP in conjunction with
Ra ≤ 0.8 µm: order code for "Measuring tube material", option SB, SE or
Ra ≤ 0.4 µm: order code for "Measuring tube material", option SC, SF
A
[mm]
B
[mm]
C
[mm]
D
[mm]
E
[mm]
L
[mm]
Endress+Hauser83
Page 84
Proline Promass F 500
L
C
D
A
B
F
E
Lap joint flange EN 1092-1, ASME B16.5, JIS B2220
Length tolerance for dimension L in mm:
+1.5 / –2.0
Lap joint flange according to EN 1092-1 Form D: PN 40
1.4301 (F304), wetted parts Alloy C22Order code for "Process connection", option DAC
DN
[mm]A[mm]B[mm]
2)
8
95654 × ⌀ 1414.54517.33700
1595654 × ⌀ 1414.54517.34040
25115854 × ⌀ 1416.56828.5444+4
401501104 × ⌀ 18218843.1560+10
501651254 × ⌀ 182310254.5719+4
802001608 × ⌀ 182913882.5848+8
1002351908 × ⌀ 2234162107.11131+4
Surface roughness (flange): Ra 3.2 to 12.5 µm
C
[mm]
D
[mm]E[mm]F[mm]L[mm]
A0022221
L
diff
[mm]
1)
1)Difference to installation length of the welding neck flange (order code for "Process connection", option
D2C)
2)DN 8 with DN 15 flanges as standard
Lap joint flange according to ASME B16.5: Class 150
1.4301 (F304), wetted parts Alloy C22Order code for "Process connection", option ADC
DN
[mm]A[mm]B[mm]
2)
8
9060.34 × ⌀ 15.71535.115.73700
C
[mm]
D
[mm]E[mm]F[mm]L[mm]
L
diff
[mm]
1)
159060.34 × ⌀ 15.71535.115.74040
2511079.44 × ⌀ 15.71650.826.74400
4012598.44 × ⌀ 15.715.973.240.95500
50150120.74 × ⌀ 19.11991.952.67150
80190152.44 × ⌀ 19.122.3127.078.08400
84Endress+Hauser
Page 85
Proline Promass F 500
Lap joint flange according to ASME B16.5: Class 150
1.4301 (F304), wetted parts Alloy C22Order code for "Process connection", option ADC
DN
[mm]A[mm]B[mm]
C
[mm]
D
[mm]E[mm]F[mm]L[mm]
L
diff
[mm]
100230190.58 × ⌀ 19.126157.2102.411270
Surface roughness (flange): Ra 3.2 to 12.5 µm
1)Difference to installation length of the welding neck flange (order code for "Process connection", option
AAC)
2)DN 8 with DN 15 flanges as standard
Lap joint flange according to ASME B16.5: Class 300
1.4301 (F304), wetted parts Alloy C22Order code for "Process connection", option AEC
DN
[mm]A[mm]B[mm]
2)
8
9566.74 × ⌀ 15.716.535.115.7376+6
C
[mm]
D
[mm]E[mm]F[mm]L[mm]
L
diff
[mm]
159566.74 × ⌀ 15.716.535.115.7406+2
2512588.94 × ⌀ 19.121.050.826.7450+10
40155114.34 × ⌀ 22.323.073.240.9564+14
501651278 × ⌀ 19.125.591.952.6717+2
80210168.38 × ⌀ 22.331.0127.078.0852.6+12.6
1002552008 × ⌀ 22.332.0157.2102.41139+12
Surface roughness (flange): Ra 3.2 to 12.5 µm
1)
1)
1)Difference to installation length of the welding neck flange (order code for "Process connection", option
ABC)
2)DN 8 with DN 15 flanges as standard
Lap joint flange according to ASME B16.5: Class 600
1.4301 (F304), wetted parts Alloy C22Order code for "Process connection", option AFC
DN
[mm]A[mm]B[mm]
2)
8
9566.74 × ⌀ 15.717.035.113.94000
C
[mm]
D
[mm]E[mm]F[mm]L[mm]
L
diff
[mm]
159566.74 × ⌀ 15.717.035.113.94200
2512588.94 × ⌀ 19.121.550.824.34900
40155114.34 × ⌀ 22.325.073.238.16000
501651278 × ⌀ 19.128.091.949.27420
80210168.38 × ⌀ 22.335.0127.073.79000
100275215.98 × ⌀ 25.444.0157.297.31167+10
Surface roughness (flange): Ra 3.2 to 12.5 µm
1)Difference to installation length of the welding neck flange (order code for "Process connection", option
ACC)
2)DN 8 with DN 15 flanges as standard
1)
Endress+Hauser85
Page 86
Proline Promass F 500
Lap joint flange JIS B2220: 20K
1.4301 (F304), wetted parts Alloy C22Order code for "Process connection", option NIC
DN
[mm]
2)
8
1595704 × ⌀ 151451154040
25125904 × ⌀ 1918.567254400
401401054 × ⌀ 1918.581405500
501551208 × ⌀ 192396507150
802001608 × ⌀ 232913280844+12
1002251858 × ⌀ 232916010011270
Surface roughness (flange): Ra 3.2 to 12.5 µm
1)Difference to installation length of the welding neck flange (order code for "Process connection", option
NEC)
2)DN 8 with DN 15 flanges as standard
A
[mm]
B
[mm]
C
[mm]
D
[mm]
E
[mm]
F
[mm]
L
[mm]
95704 × ⌀ 151451153700
L
diff
[mm]
1)
86Endress+Hauser
Page 87
Proline Promass F 500
L
A
B
Clamp connections
Tri-Clamp
Length tolerance for dimension L in mm:
+1.5 / –2.0
Tri-Clamp (½"), for pipe according to DIN 11866 series C
1.4404 (316/316L)
Order code for "Process connection", option FDW
DN
[mm]
8½25.09.5367
15½25.09.5398
3-A version available: order code for "Additional approval", option LP in conjunction with
Ra ≤ 0.8 µm: order code for "Measuring tube material", option SB, SE or
Ra ≤ 0.4 µm: order code for "Measuring tube material", option SC, SF
Clamp
[in]
A
[mm]
B
[mm]
A0015625
L
[mm]
Tri-Clamp ( ≥ 1"), for pipe according to DIN 11866 series C
1.4404 (316/316L)
Order code for "Process connection", option FTS
DN
[mm]
8150.422.1367
15150.422.1398
25150.422.1434
401½50.434.8560
50263.947.5720
80390.972.9900
1004118.997.41127
3-A version available: order code for "Additional approval", option LP in conjunction with
Ra ≤ 0.8 µm: order code for "Measuring tube material", option SB, SE or
Ra ≤ 0.4 µm: order code for "Measuring tube material", option SC, SF
Clamp
[in]
A
[mm]
B
[mm]
L
[mm]
Endress+Hauser87
Page 88
Threaded couplings
L
A
B
Thread DIN 11851, DIN11864-1, SMS 1145
Length tolerance for dimension L in mm:
+1.5 / –2.0
Thread DIN 11851, for pipe according to DIN11866, series A
1.4404 (316/316L)
Order code for "Process connection", option FMW
DN
[mm]
8Rd 34 × ¹⁄₈16367
15Rd 34 × ¹⁄₈16398
25Rd 52 × ¹⁄₆26434
40Rd 65 × ¹⁄₆38560
50Rd 78 × ¹⁄₆50720
80Rd 110 × ¹⁄₄81900
100Rd 130 × ¹⁄₄1001127
3-A version available: order code for "Additional approval", option LP in conjunction with
Ra ≤ 0.8 µm: order code for "Measuring tube material", option SB, SE
A
[in]
B
[mm]
Proline Promass F 500
A0015628
L
[mm]
Thread DIN11864-1 Form A, for pipe according to DIN11866, series A
1.4404 (316/316L)
Order code for "Process connection", option FLW
DN
[mm]
8Rd 28 × ¹⁄₈10367
15Rd 34 × ¹⁄₈16398
25Rd 52 × ¹⁄₈26434
40Rd 65 × ¹⁄₆38560
50Rd 78 × ¹⁄₆50720
80Rd 110 × ¹⁄₄81900
100Rd 130 × ¹⁄₄1001127
3-A version available: order code for "Additional approval", option LP in conjunction with
Ra ≤ 0.8 µm: order code for "Measuring tube material", option SB, SE or
Ra ≤ 0.4 µm: order code for "Measuring tube material", option SC, SF
A
[in]
B
[mm]
L
[mm]
88Endress+Hauser
Page 89
Proline Promass F 500
Thread SMS 1145
1.4404 (316/316L)
Order code for "Process connection", option SCS
DN
[mm]
8Rd 40 × ¹⁄₆22.6367
15Rd 40 × ¹⁄₆22.6398
25Rd 40 × ¹⁄₆22.6434
40Rd 60 × ¹⁄₆35.6560
50Rd 70 × ¹⁄₆48.6720
80Rd 98 × ¹⁄₆72.9900
100Rd 132 × ¹⁄₆97.61127
3-A version available: order code for "Additional approval", option LP in conjunction with
Ra ≤ 0.8 µm: order code for "Measuring tube material", option SB, SE
A
[in]
B
[mm]
L
[mm]
Endress+Hauser89
Page 90
Thread ISO 2853
L
A
B
Length tolerance for dimension L in mm:
+1.5 / –2.0
Thread ISO 2853, for pipe according to ISO 2037
1.4404 (316/316L)
Order code for "Process connection", option JSF
DN
[mm]
837.1322.6367
1537.1322.6398
2537.1322.6434
4052.6835.6560
5064.1648.6720
8091.1972.9900
100118.2197.61127
3-A version available: order code for "Additional approval", option LP in conjunction with
Ra ≤ 0.8 µm: order code for "Measuring tube material", option SB, SE or
Ra ≤ 0.4 µm: order code for "Measuring tube material", option SC, SF
1)
A
[mm]
B
[mm]
Proline Promass F 500
A0015623
L
[mm]
1)Max. thread diameter as per ISO 2853 annex A
90Endress+Hauser
Page 91
Proline Promass F 500
L
A
B
VCO
Length tolerance for dimension L in mm:
+1.5 / –2.0
8-VCO-4 (½")
1.4404 (316/316L)
Order code for "Process connection", option CVS
DN
[mm]
8AF 110.2390
A
[in]
B
[mm]
A0015624
L
[mm]
12-VCO-4 (¾")
1.4404 (316/316L)
Order code for "Process connection", option CWS
DN
[mm]
15AF 1½15.7430
A
[in]
B
[mm]
L
[mm]
Endress+Hauser91
Page 92
Accessories
DN 8 ( ")...150 (6")!⁄#DN 250 (10")
LL
45°
A
A
SW 1"
2
½" NPT
1
½" NPT
35 (1.38)ca. 75 (approx. 2.95)
1
1;2
1
1;2
Rupture disk/purge connections
Proline Promass F 500
42
1Connection nipple for purge connections: order code for "Sensor options", option CH "Purge connection"
2Connection nipple with rupture disk: order code for "Sensor option", option CA "Rupture disk"
DNAL
[mm][mm][mm]
862216
1562220
2562260
4067310
5079452
80101560
100120684
150141880
250182380
A0028914
92Endress+Hauser
Page 93
Proline Promass F 500
213 (8.4)
39 (1.5)
243 (9.6)
203 (8.0)
146 (5.75)
48 (1.9)
12 (0.47)
280 (11.0)255 (10.0)
134 (5.3)
30 (1.18)
105 (4.1)
Protective cover
A0029552
43Weather protection cover for Proline 500 – digital
44Weather protection cover for Proline 500
External WLAN antenna
The external WLAN antenna is not suitable for use in hygienic applications.
Proline 500 – digital
External WLAN antenna mounted on device
45Engineering unit mm (in)
A0029553
A0033607
Endress+Hauser93
Page 94
Proline Promass F 500
70 (2.8)
1500 (59.1)
72 (2.8)
68 (2.7)105 (4.1)
173 (6.8)
External WLAN antenna mounted with cable
The external WLAN antenna can be mounted separately from the transmitter if the transmission/
reception conditions at the transmitter mounting location are poor.
A0033606
46Engineering unit mm (in)
Proline 500
External WLAN antenna mounted on device
A0028923
47Engineering unit mm (in)
External WLAN antenna mounted with cable
The external WLAN antenna can be mounted separately from the transmitter if the transmission/
reception conditions at the transmitter mounting location are poor.
94Endress+Hauser
Page 95
Proline Promass F 500
70 (2.8)
1500 (59.1)
72 (2.8)
A0033597
48Engineering unit mm (in)
Endress+Hauser95
Page 96
Dimensions in US unitsHousing of Proline 500 – digital transmitter
N
P
F
G
A
Q
T
G
A
B
C
Q
Q
F
Non-hazardous area or hazardous area: Zone 2; Class I, Division 2
Order code for "Transmitter housing", option A "Aluminum, coated" and order code for "Integrated ISEM
electronics", option A "Sensor"
Proline Promass F 500
A0033789
A
[in]
6.579.133.157.360.940.83
F
[in]
G
[in]
N
[in]
P
[in]
Q
[in]
Order code for "Transmitter housing", option D "Polycarbonate" and order code for "Integrated ISEM
electronics", option A "Sensor"
A
[in]
6.979.213.547.760.670.87
F
[in]
G
[in]
N
[in]
P
[in]
Q
[in]
Housing of Proline 500 transmitter
Hazardous area: Zone 2; Class I, Division 2 or Zone 1; Class I, Division 1
Order code for "Transmitter housing", option A "Aluminum, coated" and order code for "Integrated ISEM
electronics", option B "Transmitter"
A
[in]
7.403.354.0612.58.545.129.41
B
[in]
C
[in]
F
[in]
G
[in]
Q
[in]
[in]
96Endress+Hauser
A0033788
T
Page 97
Proline Promass F 500
A
B
C
G
D
L
E
F
K
M
Order code for "Transmitter housing", option L "Cast, stainless" and order code for "Integrated ISEM
electronics", option B "Transmitter"
A
[in]
B
[in]
C
[in]
F
[in]
G
[in]
Q
[in]
7.403.354.0611.68.545.129.41
Sensor with connection housing
Order code for "Sensor connection housing", option A "Aluminum, coated"
DNA
1)
[in][in][in][in][in][in][in][in][in][in][in]
³⁄₈5.833.702.132.957.2810.245.350.211
½5.833.702.132.957.2810.245.350.33
15.833.702.132.957.2810.245.350.47
1½5.833.702.134.137.4611.595.350.69
25.833.702.135.557.8513.415.351.02
35.833.702.137.878.6416.525.351.59
45.833.702.13109.3719.375.352.02
65.833.702.1314.8810.225.085.352.71
105.833.702.1321.5711.9133.485.354.03
1)
B
CDE
2) 3)
F
2) 3)
GKLM
4)
4)
4)
4)
4)
4)
4)
4)
4)
T
[in]
A0033787
2.76
2.76
2.76
3.11
3.90
5.47
6.93
8.58
12.01
1)Depending on the cable gland used: values up to + 1.18 in
2)With order code for "Sensor option", option CG or order code for "Measuring tube material", option SD, SE,
SF, TH, LA: values +2.76 in
3)With order code for "Measuring tube material", option TT, TU: values +4.09 in
4)Depends on the process connection in question → 99
Order code for "Sensor connection housing", option B "Stainless"
DNA
[in][in][in][in][in][in][in][in][in][in][in]
1)
³⁄₈5.393.072.322.957.0910.045.280.211
½5.393.072.322.957.0910.045.280.33
15.393.072.322.957.0910.045.280.47
1½5.393.072.324.137.2611.45.280.69
25.393.072.325.557.6613.215.281.02
1)
B
CDE
2) 3)
F
2) 3)
GKLM
4)
4)
4)
4)
4)
2.76
2.76
2.76
3.11
3.90
Endress+Hauser97
Page 98
Proline Promass F 500
DNA
1)
1)
B
CDE
2) 3)
F
2) 3)
GKLM
[in][in][in][in][in][in][in][in][in][in][in]
35.393.072.327.878.4416.325.281.59
45.393.072.32109.1719.175.282.02
65.393.072.3214.881024.885.282.71
105.393.072.3221.5711.7133.295.284.03
4)
4)
4)
4)
5.47
6.93
8.58
12.01
1)Depending on the cable gland used: values up to +1.18 in
2)With order code for "Sensor option", option CG or order code for "Measuring tube material", option SD, SE,
SF, TH, LA: values +2.76 in
3)With order code for "Measuring tube material", option TT, TU: values +4.09 in
4)Depends on the process connection in question → 99
Order code for "Sensor connection housing", option C "Ultra-compact hygienic, stainless"
DNA
1)
[in][in][in][in][in][in][in][in][in][in][in]
³⁄₈4.882.682.202.957.0910.044.410.21
½4.882.682.202.957.0910.044.410.33
14.882.682.202.957.0910.044.410.47
1½4.882.682.204.137.2611.44.410.69
24.882.682.205.557.6613.214.411.02
34.882.682.207.878.4416.324.411.59
44.882.682.20109.1719.174.412.02
64.882.682.2014.881024.884.412.71
104.882.682.2021.5711.7133.294.414.03
1)
B
CDE
2) 3)
F
2) 3)
GKLM
4)
4)
4)
4)
4)
4)
4)
4)
4)
2.76
2.76
2.76
3.11
3.90
5.47
6.93
8.58
12.01
1)Depending on the cable gland used: values up to +1.18 in
2)With order code for "Sensor option", option CG or order code for "Measuring tube material", option SD, SE,
SF, TH, LA: values +2.76 in
3)With order code for "Measuring tube material", option TT, TU: values +4.09 in
4)Depends on the process connection in question → 99
Order code for "Sensor connection housing", option L "Cast, stainless"
DNA
1)
[in][in][in][in][in][in][in][in][in][in][in]
³⁄₈5.713.392.322.958.1911.145.350.211
½5.713.392.322.958.1911.145.350.33
15.713.392.322.958.1911.145.350.47
1½5.713.392.324.138.3712.55.350.69
25.713.392.325.558.7614.315.351.02
35.713.392.327.879.5517.425.351.59
45.713.392.321010.2820.285.352.02
65.713.392.3214.8811.125.985.352.71
105.713.392.3221.5712.8134.395.354.03
1)Depending on the cable gland used: values up to +1.18 in
2)With order code for "Sensor option", option CG or order code for "Measuring tube material", option SD, SE,
SF, TH, LA: values +2.76 in
3)With order code for "Measuring tube material", option TT, TU: values +4.09 in
4)Depends on the process connection in question → 99
1)
B
CDE
2) 3)
F
2) 3)
GKLM
4)
4)
4)
4)
4)
4)
4)
4)
4)
2.76
2.76
2.76
3.11
3.90
5.47
6.93
8.58
12.01
98Endress+Hauser
Page 99
Proline Promass F 500
C
D
A
B
E
L
Flange connections
Fixed flange ASME B16.5
A0015621
Length tolerance for dimension L in inch:
• DN ≤ 4": +0.06 / –0.08
• DN ≥ 5": +0.14
Flange according to ASME B16.5: Class 150
1.4404 (F316/F316L): order code for "Process connection", option AAS
Alloy C22: order code for "Process connection", option AAC
DN
[in]
1)
³⁄₈
½3.542.374 × ⌀0.620.440.6215.91
14.333.134 × ⌀0.620.561.0517.32
1½4.923.874 × ⌀0.620.691.6121.65
25.914.754 × ⌀0.750.752.0728.15
37.486.004 × ⌀0.750.943.0733.07
49.067.508 × ⌀0.750.944.0344.37
611.029.508 × ⌀0.8816.0755.04
1015.9414.2512 × ⌀1.01.1910.0272.13
Surface roughness (flange): Ra 125 to 250 µin
1)DN ³⁄₈" with DN ½" flanges as standard
Flange according to ASME B16.5: Class 150 with reduction in nominal diameter
1.4404 (F316/F316L)
DN
[in]
21½AHS5.914.754 × ⌀ 0.750.752.0721.65
32AJS7.4864 × ⌀ 0.750.943.0728.35
43ALS9.067.58 × ⌀ 0.750.944.0334.41
64ANS11.029.58 × ⌀ 0.8816.0745.94
86APS13.5811.758 × ⌀ 0.881.147.9857.52
Surface roughness (flange): Ra 125 to 250 µin
A
[in]
3.542.374 × ⌀ 0.620.440.6214.57
reduction
to DN
[in]
Order code for
B
[in]
"Process
connection",
option
A
[in]
C
[in]
B
[in]
D
[in]
C
[in]
E
[in]
D
[in]
E
[in]
L
[in]
L
[in]
Endress+Hauser99
Page 100
Proline Promass F 500
Flange according to ASME B16.5: Class 300
1.4404 (F316/F316L): order code for "Process connection", option ABS
Alloy C22: order code for "Process connection", option ABC
DN
[in]
1)
³⁄₈
½3.742.634 × ⌀0.620.560.6215.91
14.923.504 × ⌀0.750.691.0517.32
1½6.104.504 × ⌀0.880.811.6121.65
26.505.008 × ⌀0.750.882.0728.15
38.276.638 × ⌀0.881.123.0733.07
410.047.878 × ⌀0.881.254.0344.37
612.610.6312 × ⌀0.881.446.0755.79
1017.5215.2516 × ⌀1.121.8710.0273.35
Surface roughness (flange): Ra 125 to 250 µin
1)DN ³⁄₈" with DN ½" flanges as standard
A
[in]
B
[in]
C
[in]
D
[in]
E
[in]
3.742.634 × ⌀0.620.560.6214.57
L
[in]
Flange according to ASME B16.5: Class 300 with reduction in nominal diameter
1.4404 (F316/F316L)
DN
[in]
reduction
to DN
[in]
Order code for
"Process
connection",
A
[in]
B
[in]
C
[in]
D
[in]
E
[in]
option
21½AIS6.558 × ⌀ 0.750.882.0724.21
32AKS8.276.638 × ⌀ 0.881.123.0728.82
43AMS10.047.878 × ⌀ 0.881.254.0335.2
64AOS12.610.6312 × ⌀ 0.881.446.0746.73
86AQS14.961312 × ⌀ 11.647.9857.52
Surface roughness (flange): Ra 125 to 250 µin
Flange according to ASME B16.5: Class 600
1.4404 (F316/F316L): order code for "Process connection", option ACS
Alloy C22: order code for "Process connection", option ACC
DN
[in]
³⁄₈
1)
A
[in]
B
[in]
C
[in]
D
[in]
E
[in]
3.742.634 × ⌀0.620.810.5515.75
½3.742.634 × ⌀0.620.810.5516.54
14.923.504 × ⌀0.750.940.9619.29
1½6.104.504 × ⌀0.881.131.523.62
26.505.008 × ⌀0.751.251.9429.21
38.276.638 × ⌀0.881.52.935.43
410.838.508 × ⌀1.001.913.8345.55
613.9811.5012 × ⌀1.121.886.0757.76
1020.0817.0016 × ⌀1.382.7510.0276.61
Surface roughness (flange): Ra 125 to 250 µin
L
[in]
L
[in]
1)DN ³⁄₈" with DN ½" flanges as standard