Endress+Hauser F 500 Specifications

Page 1
TI01222D/06/EN/04.18 71402584 2018-06-01
Products Solutions Services
Technical Information
Proline Promass F 500
Coriolis flowmeter
The flowmeter offering premium accuracy and robustness, available as remote version with up to 4 I/Os
Application
• Measuring principle operates independently of physical fluid properties such as viscosity or density
• Highest measurement performance for liquids and gases under varying, demanding process conditions
Device properties
• Medium temperature: –196 to +350 °C (–320 to +662 °F)
• Nominal diameter: DN 8 to 250 (³⁄₈ to 10")
• Remote version with up to 4 I/Os
• Backlit display with touch control and WLAN access
• Standard cable between sensor and transmitter
Your benefits
• Highest process safety – immune to fluctuating and harsh environments
• Fewer process measuring points – multivariable measurement (flow, density, temperature)
• Spacesaving installation – no inlet/outlet run needs
• Full access to process and diagnostic information – numerous, freely combinable I/Os and fieldbuses
• Reduced complexity and variety – freely configurable I/O functionality
• Integrated verification – Heartbeat Technology
Page 2

Table of contents

Proline Promass F 500
About this document ........................ 4
Symbols used ................................ 4
Function and system design ................... 5
Measuring principle ............................ 5
Measuring system ............................. 7
Equipment architecture ......................... 9
Safety ..................................... 9
Input .................................... 12
Measured variable ............................ 12
Measuring range ............................. 12
Operable flow range ........................... 13
Input signal ................................ 13
Output .................................. 15
Output and input variants ....................... 15
Output signal ............................... 16
Signal on alarm .............................. 20
Ex connection data ........................... 22
Low flow cut off ............................. 23
Galvanic isolation ............................ 23
Protocol-specific data .......................... 24
Power supply ............................. 28
Terminal assignment .......................... 28
Device plugs available .......................... 29
Pin assignment, device plug ...................... 30
Supply voltage .............................. 32
Power consumption ........................... 32
Current consumption .......................... 32
Power supply failure .......................... 32
Electrical connection .......................... 33
Potential equalization ......................... 43
terminals .................................. 44
Cable entries ............................... 44
Cable specification ............................ 44
Performance characteristics .................. 49
Reference operating conditions ................... 49
Maximum measured error ....................... 49
Repeatability ............................... 51
Response time .............................. 52
Influence of ambient temperature ................. 52
Influence of medium temperature .................. 52
Influence of medium pressure .................... 53
Design fundamentals .......................... 53
Installation ............................... 54
Mounting location ............................ 54
Orientation ................................ 55
Inlet and outlet runs .......................... 56
Mounting the transmitter housing ................. 57
Special mounting instructions .................... 58
Environment .............................. 60
Ambient temperature range ..................... 60
Storage temperature .......................... 60
Climate class ............................... 60
Degree of protection .......................... 60
Vibration resistance ........................... 60
Shock resistance ............................. 61
Shock resistance ............................. 61
Interior cleaning ............................. 61
Electromagnetic compatibility (EMC) ............... 61
Process .................................. 61
Medium temperature range ...................... 61
Density ................................... 62
Pressure-temperature curves ..................... 63
Sensor housing .............................. 68
Rupture disk ................................ 69
Flow limit ................................. 69
Pressure loss ............................... 69
System pressure ............................. 69
Thermal insulation ........................... 70
Heating ................................... 70
Vibrations ................................. 71
Custody transfer measurement ............... 72
Mechanical construction .................... 73
Dimensions in SI units ......................... 73
Dimensions in US units ......................... 96
Weight .................................. 108
Materials ................................. 109
Process connections .......................... 112
Surface roughness ........................... 112
Operability .............................. 112
Operating concept ........................... 112
Languages ................................ 113
Local operation ............................. 113
Remote operation ........................... 113
Service interface ............................ 119
Network integration .......................... 120
Supported operating tools ...................... 121
HistoROM data management .................... 122
Certificates and approvals .................. 124
CE mark .................................. 124
C-Tick symbol .............................. 124
Ex approval ............................... 124
Sanitary compatibility ......................... 127
Pharmaceutical compatibility .................... 127
Functional safety ............................ 127
HART certification ........................... 127
FOUNDATION Fieldbus certification ............... 127
Certification PROFIBUS ........................ 127
EtherNet/IP certification ....................... 128
Certification PROFINET ........................ 128
Pressure Equipment Directive ................... 128
Radio approval ............................. 128
Measuring instrument approval .................. 128
Additional certification ........................ 128
2 Endress+Hauser
Page 3
Proline Promass F 500
Other standards and guidelines .................. 129
Ordering information ...................... 130
Application packages ...................... 130
Diagnostics functions ......................... 130
Heartbeat Technology ........................ 130
Concentration .............................. 131
Special density ............................. 131
Petroleum ................................ 131
OPC-UA server ............................. 131
Accessories .............................. 131
Device-specific accessories ...................... 132
Communication-specific accessories ............... 133
Service-specific accessories ..................... 134
System components .......................... 134
Supplementary documentation .............. 135
Standard documentation ....................... 135
Device-dependent additional documentation ......... 136
Registered trademarks ..................... 136
Endress+Hauser 3
Page 4

About this document

A

Symbols used Electrical symbols

Symbol Meaning
Proline Promass F 500
Direct current
Alternating current
Direct current and alternating current
Ground connection
A grounded terminal which, as far as the operator is concerned, is grounded via a grounding system.
Protective Earth (PE)
A terminal which must be connected to ground prior to establishing any other connections.
The ground terminals are situated inside and outside the device:
• Inner ground terminal: Connects the protectiv earth to the mains supply.
• Outer ground terminal: Connects the device to the plant grounding system.
Communication symbols
Symbol Meaning
Wireless Local Area Network (WLAN)
Communication via a wireless, local network.
LED
Light emitting diode is off.
LED
Light emitting diode is on.
LED
Light emitting diode is flashing.
Symbols for certain types of information
Symbol Meaning
Permitted
Procedures, processes or actions that are permitted.
Preferred
Procedures, processes or actions that are preferred.
Forbidden
Procedures, processes or actions that are forbidden.
Tip
Indicates additional information.
Reference to documentation.
Reference to page.
Reference to graphic.
Visual inspection.
4 Endress+Hauser
Page 5
Proline Promass F 500
1.
-
.
21 3
Symbols in graphics
Symbol Meaning
1, 2, 3, ... Item numbers
, 2., 3., … Series of steps
A, B, C, ... Views
A-A, B-B, C-C, ... Sections
Hazardous area
Safe area (non-hazardous area)
Flow direction

Function and system design

Measuring principle

The measuring principle is based on the controlled generation of Coriolis forces. These forces are always present in a system when both translational and rotational movements are superimposed.
Fc = 2 · ∆m (ν · ω)
Fc = Coriolis force
∆m = moving mass
ω = rotational velocity
ν = radial velocity in rotating or oscillating system
The amplitude of the Coriolis force depends on the moving mass ∆m, its velocity ν in the system and thus on the mass flow. Instead of a constant rotational velocity ω, the sensor uses oscillation.
In the sensor, two parallel measuring tubes containing flowing fluid oscillate in antiphase, acting like a tuning fork. The Coriolis forces produced at the measuring tubes cause a phase shift in the tube oscillations (see illustration):
• At zero flow (when the fluid is at a standstill) the two tubes oscillate in phase (1).
• Mass flow causes deceleration of the oscillation at the inlet of the tubes (2) and acceleration at the outlet (3).
A0028850
The phase difference (A-B) increases with increasing mass flow. Electrodynamic sensors register the tube oscillations at the inlet and outlet. System balance is ensured by the antiphase oscillation of the two measuring tubes. The measuring principle operates independently of temperature, pressure, viscosity, conductivity and flow profile.
Endress+Hauser 5
Page 6
Proline Promass F 500
Density measurement
The measuring tube is continuously excited at its resonance frequency. A change in the mass and thus the density of the oscillating system (comprising measuring tube and fluid) results in a corresponding, automatic adjustment in the oscillation frequency. Resonance frequency is thus a function of medium density. The microprocessor utilizes this relationship to obtain a density signal.
Volume measurement
Together with the measured mass flow, this is used to calculate the volume flow.
Temperature measurement
The temperature of the measuring tube is determined in order to calculate the compensation factor due to temperature effects. This signal corresponds to the process temperature and is also available as an output signal.
6 Endress+Hauser
Page 7
Proline Promass F 500
2
3
1
A
B
1
2
3

Measuring system

The measuring system consists of a transmitter and a sensor. The transmitter and sensor are mounted in physically separate locations. They are interconnected by connecting cables.
Transmitter
Two versions of the transmitter are available.
Proline 500 – digital Proline 500
For use in applications not required to meet special requirements due to ambient or operating conditions.
A Non-hazardous area or Zone 2; Class I, Division 2 B Non-hazardous area or Zone 2; Class I, Division 2 or Zone 1; Class I,
Division 1 1 Transmitter 2 Connecting cable: cable, separate, standard 3 Sensor connection housing with integrated ISEM
• Flexible and cost-effective separate installation.
• A standard cable can be used as the connecting cable.
• Electronics in the transmitter housing, ISEM (intelligent sensor
electronics module) in the sensor connection housing
• Signal transmission: digital
Order code for "Integrated ISEM electronics", option A "Sensor"
Connecting cable (can be ordered in various lengths)→  132
• Length:
– Zone 2; Class I, Division 2: max. 300 m (1 000 ft) – Zone 1; Class I, Division 1: max. 150 m (500 ft)
• Standard cable with common shield (pair-stranded)
Hazardous area
Use in: Zone 2; Class I, Division 2
Mixed installation is possible:
• Sensor: Zone 1; Class I, Division 1
• Transmitter: Zone 2; Class I, Division 2
Device versions and materials
• Transmitter housing
– Aluminum, coated: aluminum, AlSi10Mg, coated – Material: polycarbonate
• Material of window in transmitter housing
– Aluminum, coated: glass – Polycarbonate: plastic
Configuration
• External operation via 4-line, backlit, graphic local display with touch control and guided menus ("Make-it-run" wizards) for application-specific
commissioning.
• Via service interface or WLAN interface:
– Operating tools (e.g. FieldCare, DeviceCare, SmartBlue app) – Web server (access via Web browser, e.g. Microsoft Internet Explorer, Microsoft Edge)
For use in applications required to meet special requirements due to ambient or operating conditions.
Non-hazardous area or Zone 2; Class I, Division 2 or Zone 1; Class I,
Division 1 1 Transmitter with integrated ISEM 2 Connecting cable: cable, separate 3 Sensor connection housing
Application examples for sensors without electronics:
• Strong vibrations at the sensor.
• Sensor in underground installations.
• Permanent immersion of sensor in water, IP68 ingress protection.
• Electronics and ISEM (intelligent sensor electronics module) in the
transmitter housing
• Signal transmission: analog
Order code for "Integrated ISEM electronics", option B "Transmitter"
• Length: max. 20 m (65 ft)
• Cable with a common shield and individual shielded cores (3 pairs)
Use in: Zone 1; Class I, Division 1 oder Zone 2; Class I, Division 2
• Transmitter housing
– Aluminum, coated: aluminum, AlSi10Mg, coated – Cast, stainless: cast, stainless steel, 1.4409 (CF3M) similar to 316L
• Window material: glass
Endress+Hauser 7
Page 8
Sensor connection housing
Different versions of the connection housing are available.
Order code for "Sensor connection housing", option A, "Aluminum, coated": Aluminum, AlSi10Mg, coated
This device version is only available in conjunction with the Proline
500 – digital transmitter.
Order code for "Sensor connection housing", option B, "Stainless":
• Hygienic version, stainless steel 1.4301 (304)
• Optional: order code for "Sensor feature", option CC "Hygienic version, for maximum corrosion resistance": stainless steel 1.4404 (316L)
Order code for "Sensor connection housing", option C, "Ultra-compact hygienic, stainless":
• Hygienic version, stainless steel 1.4301 (304)
• Optional: order code for "Sensor feature", option CC "Hygienic version, for maximum corrosion resistance": stainless steel 1.4404 (316L)
This device version is only available in conjunction with the Proline
500 – digital transmitter.
Order code for "Sensor connection housing", option L, "Cast, stainless":
1.4409 (CF3M) similar to 316L
Proline Promass F 500
Sensor
PromassF • Excellent performance across a wide range of applications
• Simultaneous measurement of flow, volume flow, density and temperature (multivariable)
• Immune to process influences
• Materials:
A0026713
– Sensor: stainless steel, 1.4301/1.4307 (304L); optional 1.4404
(316/316L)
– Measuring tubes: stainless steel, 1.4539 (904L); 1.4404
(316/316L); Alloy C22, 2.4602 (UNS N06022)
– Process connections: stainless steel, 1.4404 (316/316L); 1.4301
(304); Alloy C22, 2.4602 (UNS N06022)
8 Endress+Hauser
Page 9
Proline Promass F 500
2
1
6
5
7
4
3
3

Equipment architecture

A0027512
 1 Possibilities for integrating measuring devices into a system
1 Control system (e.g. PLC) 2 Connecting cable (0/4 to 20 mA HART etc.) 3 Fieldbus 4 Segment coupler 5 Non-hazardous area 6 Hazardous area: Zone 2; Class I, Division 2 7 Hazardous area: Zone 1; Class I, Division 1

Safety IT security

Our warranty is valid only if the device is installed and used as described in the Operating Instructions. The device is equipped with security mechanisms to protect it against any inadvertent changes to the settings.
IT security measures, which provide additional protection for the device and associated data transfer, must be implemented by the operators themselves in line with their security standards.
Device-specific IT security
The device offers a range of specific functions to support protective measures on the operator's side. These functions can be configured by the user and guarantee greater in-operation safety if used correctly. An overview of the most important functions is provided in the following section.
Function/interface Factory setting Recommendation
Write protection via hardware write protection switch →  10
Access code (also applies for Web server login or FieldCare connection) →  10
WLAN (order option in display module)
WLAN security mode Enabled (WPA2-
WLAN passphrase (password) →  10
WLAN mode Access Point On an individual basis following risk
Web server→  10 Enabled. On an individual basis following risk
CDI-RJ45 service interface →  11 On an individual basis following risk
Not enabled. On an individual basis following risk
assessment.
Not enabled (0000).
Enabled. On an individual basis following risk
PSK)
Serial number Assign a customized access code during
Assign a customized access code during commissioning.
assessment.
Do not change.
commissioning.
assessment.
assessment.
assessment.
Endress+Hauser 9
Page 10
Proline Promass F 500
Protecting access via hardware write protection
Write access to the device parameters via the local display, Web browser or operating tool (e.g. FieldCare, DeviceCare) can be disabled via a write protection switch (DIP switch on the motherboard). When hardware write protection is enabled, only read access to the parameters is possible.
Hardware write protection is disabled when the device is delivered.
Protecting access via a password
Different passwords are available to protect write access to the device parameters or access to the device via the WLAN interface.
• User-specific access code Protect write access to the device parameters via the local display, Web browser or operating tool (e.g. FieldCare, DeviceCare). Access authorization is clearly regulated through the use of a user­specific access code.
• WLAN passphrase The network key protects a connection between an operating unit (e.g. notebook or tablet) and the device via the WLAN interface which can be ordered as an option.
• Infrastructure mode When the device is operated in infrastructure mode, the WLAN passphrase corresponds to the WLAN passphrase configured on the operator side.
User-specific access code
Write access to the device parameters via the local display, Web browser or operating tool (e.g. FieldCare, DeviceCare) can be protected by the modifiable, user-specific access code.
WLAN passphrase: Operation as WLAN access point
A connection between an operating unit (e.g. notebook or tablet) and the device via the WLAN interface, which can be ordered as an optional extra, is protected by the network key. The WLAN authentication of the network key complies with the IEEE 802.11 standard.
When the device is delivered, the network key is pre-defined depending on the device. It can be changed via the WLAN settings submenu in the WLAN passphrase parameter.
Infrastructure mode
A connection between the device and WLAN access point is protected by means of an SSID and passphrase on the system side. Please contact the relevant system administrator for access.
General notes on the use of passwords
• The access code and network key supplied with the device should be changed during commissioning.
• Follow the general rules for generating a secure password when defining and managing the access code or network key.
• The user is responsible for the management and careful handling of the access code and network key.
Access via Web server
The device can be operated and configured via a Web browser with the integrated Web server. The connection is via the service interface (CDI-RJ45) or the WLAN interface. For device versions with the EtherNet/IP and PROFINET communication protocols, the connection can also be established via the terminal connection for signal transmission with EtherNet/IP or PROFINET (RJ45 connector).
The Web server is enabled when the device is delivered. The Web server can be disabled if necessary (e.g. after commissioning) via the Web server functionality parameter.
The device and status information can be hidden on the login page. This prevents unauthorized access to the information.
For detailed information on device parameters, see: The "Description of Device Parameters" document →  135
Access via OPC-UA
The "OPC UA Server" application package is available in the device version with the HART communication protocol →  131.
10 Endress+Hauser
Page 11
Proline Promass F 500
The device can communicate with OPC UA clients using the "OPC UA Server" application package.
The OPC UA server integrated in the device can be accessed via the WLAN access point using the WLAN interface - which can be ordered as an optional extra - or the service interface (CDI- RJ45) via Ethernet network. Access rights and authorization as per separate configuration.
The following Security Modes are supported as per the OPC UA Specification (IEC 62541):
• None
• Basic128Rsa15 – signed
• Basic128Rsa15 – signed and encrypted
Access via service interface (CDI-RJ45)
The device can be connected to a network via the service interface (CDI-RJ45). Device-specific functions guarantee the secure operation of the device in a network.
The use of relevant industrial standards and guidelines that have been defined by national and international safety committees, such as IEC/ISA62443 or the IEEE, is recommended. This includes organizational security measures such as the assignment of access authorization as well as technical measures such as network segmentation.
The device can be integrated in a ring topology. The device is integrated via the terminal connection for signal transmission (output 1) and the connection to the service interface (CDI­RJ45) →  117.
Endress+Hauser 11
Page 12

Input

Measured variable Direct measured variables

• Mass flow
• Density
• Temperature
Calculated measured variables
• Volume flow
• Corrected volume flow
• Reference density

Measuring range Measuring ranges for liquids

Proline Promass F 500
DN Measuring range full scale values 
min(F)
to 
[mm] [in] [kg/h] [lb/min]
8 ³⁄₈ 0 to 2 000 0 to 73.50
15 ½ 0 to 6 500 0 to 238.9
25 1 0 to 18 000 0 to 661.5
40 0 to 45 000 0 to 1 654
50 2 0 to 70 000 0 to 2 573
80 3 0 to 180 000 0 to 6 615
100 4 0 to 350 000 0 to 12 860
150 6 0 to 800 000 0 to 29 400
250 10 0 to 2 200 000 0 to 80 850
Measuring range for gases
The full scale value depends on the density and the sound velocity of the gas used and can be calculated with the formula below: 
= minimum (
max(G)
max(G)
max(F)
< 
ρ
max(G)
G
max(F)
x Constant dependent on nominal diameter
c
G
d
i
· ρG : x ; ρG · cG · π/2 · (di)2 · 3600)
max(F)
Maximum full scale value for gas [kg/h]
Maximum full scale value for liquid [kg/h]
can never be greater than 
max(G)
Gas density in [kg/m³] at operating conditions
Sound velocity (gas) [m/s]
Measuring tube internal diameter [m]
max(F)
max(F)
DN x
[mm] [in] [kg/m3]
8 ³⁄₈ 60
15 ½ 80
25 1 90
40 90
50 2 90
80 3 110
12 Endress+Hauser
Page 13
Proline Promass F 500
DN x
[mm] [in] [kg/m3]
100 4 130
150 6 200
250 10 200
To calculate the measuring range, use the Applicator sizing tool →  134
Calculation example for gas
• Sensor: Promass F, DN 50
• Gas: Air with a density of 60.3 kg/m³ (at 20 °C and 50 bar)
• Measuring range (liquid): 70 000 kg/h
• x = 90 kg/m³ (for Promass F, DN 50) Maximum possible full scale value:
max(G)
= 
· ρG : x = 70 000 kg/h · 60.3 kg/m³ : 90 kg/m³ = 46 900 kg/h
max(F)
Recommended measuring range
"Flow limit" section →  69

Operable flow range

Over 1000 : 1.
Flow rates above the preset full scale value do not override the electronics unit, with the result that the totalizer values are registered correctly.

Input signal Input and output versions

→  15
External measured values
To increase the accuracy of certain measured variables or to calculate the corrected volume flow for gases, the automation system can continuously write different measured values to the measuring device:
• Operating pressure to increase accuracy (Endress+Hauser recommends the use of a pressure measuring device for absolute pressure, e.g. Cerabar M or Cerabar S)
• Medium temperature to increase accuracy (e.g. iTEMP)
• Reference density for calculating the corrected volume flow for gases
Various pressure transmitters and temperature measuring devices can be ordered from Endress +Hauser: see "Accessories" section →  134
It is recommended to read in external measured values to calculate the corrected volume flow.
HART protocol
The measured values are written from the automation system to the measuring device via the HART protocol. The pressure transmitter must support the following protocol-specific functions:
• HART protocol
• Burst mode
Current input
The measured values are written from the automation system to the measuring device via the current input →  14.
Digital communication
The measured values can be written from the automation system to the measuring via:
• FOUNDATION Fieldbus
• PROFIBUS DP
• PROFIBUS PA
• Modbus RS485
• EtherNet/IP
• PROFINET
Endress+Hauser 13
Page 14
Current input 0/4 to 20 mA
Current input 0/4 to 20 mA (active/passive)
Current span • 4 to 20 mA (active)
• 0/4 to 20 mA (passive)
Resolution 1 µA
Voltage drop Typically: 0.6 to 2 V for 3.6 to 22 mA (passive)
Maximum input voltage ≤ 30 V (passive)
Open-circuit voltage ≤ 28.8 V (active)
Possible input variables • Pressure
• Temperature
• Density
Status input
Maximum input values • DC –3 to 30 V
• If status input is active (ON): Ri >3 kΩ
Response time Adjustable: 5 to 200 ms
Input signal level • Low signal: DC –3 to +5 V
• High signal: DC 12 to 30 V
Assignable functions • Off
• Reset the individual totalizers separately
• Reset all totalizers
• Flow override
Proline Promass F 500
14 Endress+Hauser
Page 15
Proline Promass F 500

Output

Output and input variants

Depending on the option selected for output/input 1, different options are available for the other outputs and inputs. Only one option can be selected for each output/input 1 to 4. The table must be read vertically (↓).
Example: If the option BA "4–20 mA HART" was selected for output/input 1, one of the options A, B, D, E, F, H, I or J is available for output 2, and one of the options A, B, D, E, F, H, I or J is available for output 3 and 4.
Order code for "Output; input 1" (020) → Possible options
Current output 4 to 20 mA HART BA
Current output 4 to 20 mA HART Ex i CA
FOUNDATION Fieldbus SA
FOUNDATION Fieldbus Ex i TA
PROFIBUS DP LA
PROFIBUS PA GA
PROFIBUS PA Ex i HA
Modbus RS485 MA
EtherNet/IP 2-port switch integrated NA
PROFINET 2-port switch integrated RA
Order code for "Output; input 2" (021) →
Not assigned A A A A A A A A A A
Current output 0/4 to 20 mA B B B B B B B
Current output 0/4 to 20 mA (Ex i) C C C
User configurable input/output
Pulse/frequency/switch output E E E E E E E
Double pulse output
Pulse/frequency/switch output (Ex i) G G G
Relay output H H H H H H H
Current input 0/4 to 20 mA I I I I I I I
Status input J J J J J J J
Order code for "Output; input 3" (022), "Output; input 4" (023)
Not assigned A A A A A A A A A A
Current output 0/4 to 20 mA B B B B B
Current output 0/4 to 20 mA (Ex i) C
User configurable input/output D D D D D
Pulse/frequency/switch output E E E E E
Double pulse output (slave)
Pulse/frequency/switch output (Ex i) G
Relay output H H H H H
Current input 0/4 to 20 mA I I I I I
Status input J J J J J
2)
1)
3)
4)
D D D D D D D
F F
F F
1) A specific input or output can be assigned to a user configurable input/output →  19.
2) If double pulse output (F) is selected for output/input 2 (021), only the double pulse output (F) option is available for selection for output/input 3 (022).
3) The order code for "Output; input 4" (023) is only available for the Proline 500 – digital transmitter.
4) The double pulse output (F) option is not available for input/output 4.
Endress+Hauser 15
Page 16

Output signal HART current output

Current output 4 to 20 mA HART
Current span Can be set to: 4 to 20 mA (active/passive)
Open-circuit voltage DC 28.8 V (active)
Maximum input voltage DC 30 V (passive)
Load 250 to 700 Ω
Resolution 0.38 µA
Damping Configurable: 0.07 to 999 s
Assignable measured variables
Proline Promass F 500
Ex-i, passive
• Mass flow
• Volume flow
• Corrected volume flow
• Density
• Reference density
• Temperature
• Electronic temperature
• Oscillation frequency 0
• Oscillation damping 0
• Signal asymmetry
• Exciter current 0
The range of options increases if the measuring device has one or more
application packages.
PROFIBUS PA
PROFIBUS PA In accordance with EN 50170 Volume 2, IEC 61158-2 (MBP), galvanically
isolated
Data transmission 31.25 kbit/s
Current consumption 10 mA
16 mA
Permitted supply voltage 9 to 32 V
Bus connection With integrated reverse polarity protection
PROFIBUS DP
Signal encoding NRZ code
Data transfer 9.6 kBaud…12 MBaud
EtherNet/IP
Standards In accordance with IEEE 802.3
PROFINET
Standards In accordance with IEEE 802.3
FOUNDATION Fieldbus
FOUNDATION Fieldbus H1, IEC 61158-2, galvanically isolated
Data transfer 31.25 kbit/s
16 Endress+Hauser
Page 17
Proline Promass F 500
Current consumption 10 mA
Permitted supply voltage 9 to 32 V
Bus connection With integrated reverse polarity protection
Modbus RS485
Physical interface RS485 in accordance with EIA/TIA-485 standard
Terminating resistor Integrated, can be activated via DIP switches
Current output 0/4 to 20 mA
Current output 0/4 to 20 mA
Maximum output values 22.5 mA
Current span Can be set to:
• 4 to 20 mA (active)
• 0/4 to 20 mA (passive)
Ex-i, passive
Open-circuit voltage DC 28.8 V (active)
Maximum input voltage DC 30 V (passive)
Load 0 to 700 Ω
Resolution 0.38 µA
Damping Adjustable: 0.07 to 999 s
Assignable measured variables
• Mass flow
• Volume flow
• Corrected volume flow
• Density
• Reference density
• Temperature
• Electronic temperature
• Oscillation frequency 0
• Oscillation damping 0
• Signal asymmetry
• Exciter current 0
The range of options increases if the measuring device has one or more
application packages.
Pulse/frequency/switch output
Function Can be set to pulse, frequency or switch output
Version Open collector
Can be set to:
• Active
• Passive
Ex-i, passive
Maximum input values DC 30 V, 250 mA (passive)
Open-circuit voltage DC 28.8 V (active)
Voltage drop For 22.5 mA: ≤ DC 2 V
Pulse output
Maximum input values DC 30 V, 250 mA (passive)
Maximum output current 22.5 mA (active)
Endress+Hauser 17
Page 18
Open-circuit voltage DC 28.8 V (active)
Pulse width Adjustable: 0.05 to 2 000 ms
Maximum pulse rate 10 000 Impulse/s
Pulse value Adjustable
Assignable measured variables
Frequency output
Maximum input values DC 30 V, 250 mA (passive)
Maximum output current 22.5 mA (active)
Open-circuit voltage DC 28.8 V (active)
Output frequency Adjustable: end value frequency 2 to 10 000 Hz (f
Damping Adjustable: 0 to 999 s
Pulse/pause ratio 1:1
Assignable measured variables
Switch output
Maximum input values DC 30 V, 250 mA (passive)
Open-circuit voltage DC 28.8 V (active)
Switching behavior Binary, conductive or non-conductive
Switching delay Adjustable: 0 to 100 s
Number of switching cycles
Assignable functions • Off
• Mass flow
• Volume flow
• Corrected volume flow
• Mass flow
• Volume flow
• Corrected volume flow
• Density
• Reference density
• Temperature
• Electronic temperature
• Oscillation frequency 0
• Oscillation damping 0
• Signal asymmetry
• Exciter current 0
The range of options increases if the measuring device has one or more
application packages.
Unlimited
• On
• Diagnostic behavior
• Limit value – Mass flow – Volume flow – Corrected volume flow – Density – Reference density – Temperature – Totalizer 1-3
• Flow direction monitoring
• Status – Partially filled pipe detection – Low flow
The range of options increases if the measuring device has one or more
application packages.
Proline Promass F 500
= 12 500 Hz)
max
18 Endress+Hauser
Page 19
Proline Promass F 500
Double pulse output
Function Double pulse
Version Open collector
Can be set to:
• Active
• Passive
Maximum input values DC 30 V, 250 mA (passive)
Open-circuit voltage DC 28.8 V (active)
Voltage drop For 22.5 mA: ≤ DC 2 V
Output frequency Adjustable: 0 to 1 000 Hz
Damping Adjustable: 0 to 999 s
Pulse/pause ratio 1:1
Assignable measured variables
• Mass flow
• Volume flow
• Corrected volume flow
• Density
• Reference density
• Temperature
The range of options increases if the measuring device has one or more
application packages.
Relay output
Function Switch output
Version Relay output, galvanically isolated
Switching behavior Can be set to:
• NO (normally open), factory setting
• NC (normally closed)
Maximum switching capacity (passive)
Assignable functions • Off
• DC 30 V, 0.1 A
• AC 30 V, 0.5 A
• On
• Diagnostic behavior
• Limit value – Mass flow – Volume flow – Corrected volume flow – Density – Reference density – Temperature – Totalizer 1-3
• Flow direction monitoring
• Status – Partially filled pipe detection – Low flow
The range of options increases if the measuring device has one or more
application packages.
User configurable input/output
One specific input or output is assigned to a user-configurable input/output (configurable I/O)
during device commissioning.
The following inputs and outputs are available for assignment:
• Choice of current output: 4 to 20 mA (active), 0/4 to 20 mA (passive)
• Pulse/frequency/switch output
• Choice of current input: 4 to 20 mA (active), 0/4 to 20 mA (passive)
• Status input
Endress+Hauser 19
Page 20
Proline Promass F 500
The technical values correspond to those of the inputs and outputs described in this section.

Signal on alarm

Depending on the interface, failure information is displayed as follows:
HART current output
Device diagnostics Device condition can be read out via HART Command 48
PROFIBUS PA
Status and alarm messages
Error current FDE (Fault Disconnection Electronic)
Diagnostics in accordance with PROFIBUS PA Profile 3.02
0 mA
PROFIBUS DP
Status and alarm messages
Diagnostics in accordance with PROFIBUS PA Profile 3.02
EtherNet/IP
Device diagnostics Device condition can be read out in Input Assembly
PROFINET
Device diagnostics According to "Application Layer protocol for decentralized periphery", Version 2.3
FOUNDATION Fieldbus
Status and alarm messages
Error current FDE (Fault Disconnection Electronic)
Diagnostics in accordance with FF-891
0 mA
Modbus RS485
Failure mode Choose from:
• NaN value instead of current value
• Last valid value
Current output 0/4 to 20 mA
4 to 20 mA
Failure mode Choose from:
• 4 to 20 mA in accordance with NAMUR recommendation NE 43
• 4 to 20 mA in accordance with US
• Min. value: 3.59 mA
• Max. value: 22.5 mA
• Freely definable value between: 3.59 to 22.5 mA
• Actual value
• Last valid value
20 Endress+Hauser
Page 21
Proline Promass F 500
0 to 20 mA
Failure mode Choose from:
• Maximum alarm: 22 mA
• Freely definable value between: 0 to 20.5 mA
Pulse/frequency/switch output
Pulse output
Failure mode Choose from:
• Actual value
• No pulses
Frequency output
Failure mode Choose from:
• Actual value
• 0 Hz
• Defined value (f
Switch output
Failure mode Choose from:
• Current status
• Open
• Closed
2 to 12 500 Hz)
max
Relay output
Failure mode Choose from:
• Current status
• Open
• Closed
Local display
Plain text display With information on cause and remedial measures
Backlight Red backlighting indicates a device error.
Status signal as per NAMUR recommendation NE 107
Interface/protocol
• Via digital communication: – HART protocol – FOUNDATION Fieldbus – PROFIBUS PA – PROFIBUS DP – Modbus RS485 – EtherNet/IP – PROFINET
• Via service interface – CDI-RJ45 service interface – WLAN interface
Plain text display With information on cause and remedial measures
Additional information on remote operation →  113
Endress+Hauser 21
Page 22
Web server
Plain text display With information on cause and remedial measures
Light emitting diodes (LED)
Status information Status indicated by various light emitting diodes

Ex connection data Safety-related values

Proline Promass F 500
The following information is displayed depending on the device version:
• Supply voltage active
• Data transmission active
• Device alarm/error has occurred
• EtherNet/IP network available
• EtherNet/IP connection established
• PROFINET network available
• PROFINET connection established
• PROFINET blinking feature
Order code for
Output type Safety-related values
"Output; input 1"
Option BA Current output
4 to 20 mA HART
UN = 30 V UM = 250 V
Option GA PROFIBUS PA UN = 30 V
UM = 250 V
Option LA PROFIBUS DP UN = 30 V
UM = 250 V
Option MA Modbus RS485 UN = 30 V
UM = 250 V
Option SA FOUNDATION Fieldbus UN = 30 V
UM = 250 V
Option NA EtherNet/IP UN = 30 V
UM = 250 V
Option RA PROFINET UN = 30 V
UM = 250 V
Order code for "Output; input 2"; "Output; input 3"
Output type Safety-related values
Output; input 2 Output; input 3 Output; input
"Output; input 4"
24 (+) 25 (–) 22 (+) 23 (–) 20 (+) 21 (–)
Option B Current output
4 to 20 mA
Option D User configurable input/
output
Option E Pulse/frequency/switch
output
Option F Double pulse output UN = 30 V
UN = 30 V UM = 250 V
UN = 30 V UM = 250 V
UN = 30 V UM = 250 V
DC
DC
DC
DC
UM = 250 V
Option H Relay output UN = 30 V
DC
IN =100 mADC/500 mA UM = 250 V
"Output; input 1"
26 (+) 27 (–)
DC
AC
DC
AC
DC
AC
DC
AC
DC
AC
DC
AC
DC
AC
AC
AC
AC
AC
AC
AC
1)
4
22 Endress+Hauser
Page 23
Proline Promass F 500
Order code for "Output; input 2"; "Output; input 3"
Output type Safety-related values
Output; input 2 Output; input 3 Output; input
"Output; input 4"
24 (+) 25 (–) 22 (+) 23 (–) 20 (+) 21 (–)
Option I Current input
4 to 20 mA
Option J Status input UN = 30 V
UN = 30 V UM = 250 V
UM = 250 V
DC
AC
DC
AC
1) The order code "Output; input 4" is only available for the Proline 500 – digital transmitter.
Intrinsically safe values
Order code for "Output; input 1"
Option CA Current output
Option HA PROFIBUS PA Ex i Ex ia
Option TA FOUNDATION Fieldbus
Output type Intrinsically safe values
"Output; input 1"
26 (+) 27 (–)
Ui = 30 V
4 to 20 mA HART Ex i
li = 100 mA Pi = 1.25 W Li = 0 Ci = 0
Ex i
1)
Ui = 30 V li = 570 mA Pi = 8.5 W Li = 10 µH Ci = 5 nF
1)
Ex ia
Ui = 30 V li = 570 mA Pi = 8.5 W Li = 10 µH Ci = 5 nF
2)
Ex ic
Ui = 32 V li = 570 mA Pi = 8.5 W Li = 10 µH Ci = 5 nF
2)
Ex ic
Ui = 32 V li = 570 mA Pi = 8.5 W Li = 10 µH Ci = 5 nF
1)
4

Low flow cut off

Galvanic isolation

1) Only available for the Zone 1; Class I, Division 1 version
2) Only available for the Zone 2; Class I, Division 2 version and only for the Proline 500 – digital transmitter
Order code for "Output; input 2"; "Output; input 3" "Output; input 4"
Option C Current output
Output type Intrinsically safe values or NIFW values
Output; input 2 Output; input 3 Output; input 4
24 (+) 25 (–) 22 (+) 23 (–) 20 (+) 21 (–)
Ui = 30 V
4 to 20 mA Ex i
li = 100 mA Pi = 1.25 W Li = 0 Ci = 0
Option G Pulse/frequency/switch
output Ex i
Ui = 30 V li = 100 mA Pi = 1.25 W Li = 0 Ci = 0
1) The order code "Output; input 4" is only available for the Proline 500 – digital transmitter.
The switch points for low flow cut off are user-selectable.
The outputs are galvanically isolated from one another and from earth (PE).
1)
Endress+Hauser 23
Page 24

Protocol-specific data HART

Manufacturer ID 0x11
Device type ID 0x3B
HART protocol revision 7
Device description files (DTM, DD)
HART load Min. 250 Ω
System integration Information on system integration: Operating Instructions →  135.
PROFIBUS PA
Manufacturer ID 0x11
Ident number 0x156D
Profile version 3.02
Device description files (GSD, DTM, DD)
Supported functions • Identification & Maintenance
Configuration of the device address
Compatibility with earlier model
System integration Information regarding system integration: Operating Instructions →  135.
Proline Promass F 500
Information and files under:
www.endress.com
• Measured variables via HART protocol
• Burst Mode functionality
Information and files under:
www.endress.com
www.profibus.org
Simplest device identification on the part of the control system and nameplate
• PROFIBUS upload/download Reading and writing parameters is up to ten times faster with PROFIBUS upload/download
• Condensed status Simplest and self-explanatory diagnostic information by categorizing diagnostic messages that occur
• DIP switches on the I/O electronics module
• Local display
• Via operating tools (e.g. FieldCare)
If the device is replaced, the measuring device Promass 500 supports the compatibility of the cyclic data with previous models. It is not necessary to adjust the engineering parameters of the PROFIBUS network with the Promass 500 GSD file.
Earlier models:
• Promass 80 PROFIBUS PA – ID No.: 1528 (hex) – Extended GSD file: EH3x1528.gsd – Standard GSD file: EH3_1528.gsd
• Promass 83 PROFIBUS PA – ID No.: 152A (hex) – Extended GSD file: EH3x152A.gsd – Standard GSD file: EH3_152A.gsd
Description of the function scope of compatibility:
Operating Instructions →  135.
• Cyclic data transmission
• Block model
• Description of the modules
PROFIBUS DP
Manufacturer ID 0x11
Ident number 0x156F
24 Endress+Hauser
Page 25
Proline Promass F 500
Profile version 3.02
Device description files (GSD, DTM, DD)
Supported functions • Identification & Maintenance
Configuration of the device address
System integration Information regarding system integration: Operating Instructions →  135.
Information and files under:
www.endress.com On the product page for the device: Documents/Software → Device drivers
www.profibus.org
Simplest device identification on the part of the control system and nameplate
• PROFIBUS upload/download Reading and writing parameters is up to ten times faster with PROFIBUS upload/download
• Condensed status Simplest and self-explanatory diagnostic information by categorizing diagnostic messages that occur
• DIP switches on the I/O electronics module
• Via operating tools (e.g. FieldCare)
• Cyclic data transmission
• Block model
• Description of the modules
EtherNet/IP
Protocol • The CIP Networks Library Volume 1: Common Industrial Protocol
• The CIP Networks Library Volume 2: EtherNet/IP Adaptation of CIP
Communication type • 10Base-T
• 100Base-TX
Device profile Generic device (product type: 0x2B)
Manufacturer ID 0x11
Device type ID 0x103B
Baud rates Automatic ¹⁰⁄₁₀₀ Mbit with half-duplex and full-duplex detection
Polarity Auto-polarity for automatic correction of crossed TxD and RxD pairs
Supported CIP connections Max. 3 connections
Explicit connections Max. 6 connections
I/O connections Max. 6 connections (scanner)
Configuration options for measuring device
Configuration of the EtherNet interface
Configuration of the device address
Device Level Ring (DLR) Yes
System integration Information regarding system integration: Operating Instructions
• DIP switches on the electronics module for IP addressing
• Manufacturer-specific software (FieldCare)
• Add-on Profile Level 3 for Rockwell Automation control systems
• Web browser
• Electronic Data Sheet (EDS) integrated in the measuring device
• Speed: 10 MBit, 100 MBit, auto (factory setting)
• Duplex: half-duplex, full-duplex, auto (factory setting)
• DIP switches on the electronics module for IP addressing (last octet)
• DHCP
• Manufacturer-specific software (FieldCare)
• Add-on Profile Level 3 for Rockwell Automation control systems
• Web browser
• EtherNet/IP tools, e.g. RSLinx (Rockwell Automation)
→  135.
• Cyclic data transmission
• Block model
• Input and output groups
Endress+Hauser 25
Page 26
Proline Promass F 500
PROFINET
Protocol "Application layer protocol for decentral device periphery and distributed
automation", version 2.3
Communication type 100 MBit/s
Conformity class Conformance Class B
Netload Class Netload Class II
Baud rates Automatic 100 Mbit/s with full-duplex detection
Cycle times From 8 ms
Polarity Auto-polarity for automatic correction of crossed TxD and RxD pairs
Media Redundancy Protocol (MRP)
Device profile Application interface identifier 0xF600
Manufacturer ID 0x11
Device type ID 0x843B
Device description files (GSD, DTM, DD)
Supported connections • 1 x AR (IO Controller AR)
Configuration options for measuring device
Configuration of the device name
Supported functions • Identification & Maintenance
System integration Information regarding system integration: Operating Instructions →  135.
Yes
Generic device
Information and files under:
www.endress.com On the product page for the device: Documents/Software → Device drivers
www.profibus.org
• 1 x AR (IO-Supervisor Device AR connection allowed)
• 1 x Input CR (Communication Relation)
• 1 x Output CR (Communication Relation)
• 1 x Alarm CR (Communication Relation)
• DIP switches on the electronics module, for device name assignment (last part)
• Manufacturer-specific software (FieldCare, DeviceCare)
• Web browser
• Device master file (GSD), can be read out via the integrated Web server of the measuring device
• DIP switches on the electronics module, for device name assignment (last part)
• DCP protocol
• Process Device Manager (PDM)
• Integrated Web server
Simple device identification via: – Control system – Nameplate
• Measured value status The process variables are communicated with a measured value status
• Blinking feature via the onsite display for simple device identification and assignment
• Device operation via operating tools (e.g. FieldCare, DeviceCare, SIMATIC PDM)
• Cyclic data transmission
• Overview and description of the modules
• Status coding
• Startup configuration
• Factory setting:
FOUNDATION Fieldbus
Manufacturer ID 0x452B48 (hex)
Ident number 0x103B (hex)
26 Endress+Hauser
Page 27
Proline Promass F 500
Device revision 1
DD revision Information and files under:
CFF revision
Interoperability Test Kit (ITK) Version 6.2.0
ITK Test Campaign Number Information:
Link Master capability (LAS) Yes
Choice of "Link Master" and "Basic Device"
Node address Factory setting: 247 (0xF7)
Supported functions The following methods are supported:
Virtual Communication Relationships (VCRs)
Number of VCRs 44
Number of link objects in VFD 50
Permanent entries 1
Client VCRs 0
Server VCRs 10
Source VCRs 43
Sink VCRs 0
Subscriber VCRs 43
Publisher VCRs 43
Device Link Capabilities
Slot time 4
Min. delay between PDU 8
Max. response delay 16
System integration Information regarding system integration: Operating Instructions →  135.
www.endress.com
www.fieldbus.org
www.endress.com
www.fieldbus.org
Yes Factory setting: Basic Device
• Restart
• ENP Restart
• Diagnostic
• Set to OOS
• Set to AUTO
• Read trend data
• Read event logbook
• Cyclic data transmission
• Description of the modules
• Execution times
• Methods
Modbus RS485
Protocol Modbus Applications Protocol Specification V1.1
Response times • Direct data access: typically 25 to 50 ms
• Auto-scan buffer (data range): typically 3 to 5 ms
Device type Slave
Slave address range 1 to 247
Broadcast address range 0
Endress+Hauser 27
Page 28
Proline Promass F 500
Function codes • 03: Read holding register
• 04: Read input register
• 06: Write single registers
• 08: Diagnostics
• 16: Write multiple registers
• 23: Read/write multiple registers
Broadcast messages Supported by the following function codes:
• 06: Write single registers
• 16: Write multiple registers
• 23: Read/write multiple registers
Supported baud rate • 1 200 BAUD
• 2 400 BAUD
• 4 800 BAUD
• 9 600 BAUD
• 19 200 BAUD
• 38 400 BAUD
• 57 600 BAUD
• 115 200 BAUD
Data transfer mode • ASCII
• RTU
Data access Each device parameter can be accessed via Modbus RS485.
For Modbus register information
Compatibility with earlier model
System integration Information on system integration: Operating Instructions →  135.
If the device is replaced, the measuring device Promass 500 supports the compatibility of the Modbus registers for the process variables and the diagnostic information with the previous model Promass 83. It is not necessary to change the engineering parameters in the automation system.
Description of the function scope of compatibility:
Operating Instructions →  135.
• Modbus RS485 information
• Function codes
• Register information
• Response time
• Modbus data map

Power supply

Terminal assignment Transmitter: supply voltage, input/outputs

HART
Supply voltage Input/output
1
1 (+) 2 (–) 26 (+) 27 (–) 24 (+) 25 (–) 22 (+) 23 (–) 20 (+) 21 (–)
The terminal assignment depends on the specific device version ordered →  15.
Input/output
Input/output
2
3
Input/output
4
FOUNDATION Fieldbus
Supply voltage Input/output
1
1 (+) 2 (–) 26 (A) 27 (B) 24 (+) 25 (–) 22 (+) 23 (–) 20 (+) 21 (–)
The terminal assignment depends on the specific device version ordered →  15.
Input/output
2
Input/output
3
Input/output
4
28 Endress+Hauser
Page 29
Proline Promass F 500
PROFIBUS PA
Supply voltage Input/output
1
1 (+) 2 (–) 26 (B) 27 (A) 24 (+) 25 (–) 22 (+) 23 (–) 20 (+) 21 (–)
The terminal assignment depends on the specific device version ordered →  15.
Input/output
2
Input/output
3
Input/output
4
PROFIBUS DP
Supply voltage Input/output
1
1 (+) 2 (–) 26 (B) 27 (A) 24 (+) 25 (–) 22 (+) 23 (–) 20 (+) 21 (–)
The terminal assignment depends on the specific device version ordered →  15.
Input/output
2
Input/output
3
Input/output
4
Modbus RS485
Supply voltage Input/output
1
1 (+) 2 (–) 26 (B) 27 (A) 24 (+) 25 (–) 22 (+) 23 (–) 20 (+) 21 (–)
The terminal assignment depends on the specific device version ordered →  15.
Input/output
2
Input/output
3
Input/output
4
PROFINET
Supply voltage Input/output
1
1 (+) 2 (–) PROFINET
(RJ45 connector)
Input/output
2
24 (+) 25 (–) 22 (+) 23 (–) 20 (+) 21 (–)
The terminal assignment depends on the specific device version
Input/output
3
ordered →  15.
Input/output
4

Device plugs available

EtherNet/IP
Supply voltage Input/output
1
1 (+) 2 (–) EtherNet/IP
(RJ45 connector)
Input/output
2
24 (+) 25 (–) 22 (+) 23 (–) 20 (+) 21 (–)
The terminal assignment depends on the specific device version
Input/output
3
ordered →  15.
Input/output
4
Transmitter and sensor connection housing: connecting cable
The sensor and transmitter, which are mounted in separate locations, are interconnected by a connecting cable. The cable is connected via the sensor connection housing and the transmitter housing.
Terminal assignment and connection of the connecting cable:
• Proline 500 – digital→  33
• Proline 500 →  34
Device plugs may not be used in hazardous areas!
Device plugs for fieldbus systems:
Order code for "Input; output 1"
• Option SA "FOUNDATION Fieldbus" →  30
• Option GA "PROFIBUS PA" →  30
• Option RA "PROFINET" →  30
• Option NA "EtherNet/IP" →  30
Device plug for connecting to the service interface:
Order code for "Accessory mounted" option NB, adapter RJ45 M12 (service interface) →  31
Endress+Hauser 29
Page 30
Order code for "Input; output 1", option SA "FOUNDATION Fieldbus"
1
2
4
3
Order code for Cable entry/connection →  34
"Electrical connection" 2 3
M, 3, 4, 5 7/8" connector
Order code for "Input; output 1", option GA "PROFIBUS PA"
Order code for Cable entry/connection →  34
"Electrical connection" 2 3
L, N, P, U Connector M12 × 1
Order code for "Input; output 1", option RA "PROFINET"
Order code for Cable entry/connection →  34
"Electrical connection" 2 3
L, N, P, U Connector M12 × 1
1) 2)
1) 2)
1) 2)
R
, S
, T
, V
1) 2)
Connector M12 × 1 Connector M12 × 1
Proline Promass F 500
1) Cannot be combined with an external WLAN antenna (order code for "Enclosed accessories", option P8) of an RJ45 M12 adapter for the service interface (order code for "Accessories mounted", option NB) or of the remote display and operating module DKX001.
2) Suitable for integrating the device in a ring topology.
Order code for "Input; output 1", option NA "EtherNet/IP"
Order code for Cable entry/connection →  34
"Electrical connection" 2 3
L, N, P, U Connector M12 × 1
1) 2)
1) 2)
1) 2)
R
, S
, T
, V
1) 2)
Connector M12 × 1 Connector M12 × 1
1) Cannot be combined with an external WLAN antenna (order code for "Enclosed accessories", option P8) of an RJ45 M12 adapter for the service interface (order code for "Accessories mounted", option NB) or of the remote display and operating module DKX001
2) Suitable for integrating the device in a ring topology.
Order code for "Accessory mounted", option NB "Adapter RJ45 M12 (service interface)"
Order code Cable entry/coupling →  34
"Accessory mounted" Cable entry
2
NB Plug M12 × 1
Cable entry
3

Pin assignment, device plug FOUNDATION Fieldbus

Pin Assignment Coding Plug/socket
1 + Signal + A Plug
2 - Signal –
3 Grounding
4 Not assigned
30 Endress+Hauser
Page 31
Proline Promass F 500
1
2
4
3
3
2
4
3
2
4
3
2
4
PROFIBUS PA
PROFINET
Pin Assignment Coding Plug/socket
1 + PROFIBUS PA + A Plug
2 Grounding
3 - PROFIBUS PA –
4 Not assigned
Pin Assignment
1 + TD +
2 + RD +
3 - TD –
4 - RD –
A0032047
Coding Plug/socket
D Socket
Recommended plug:
• Binder, series 763, part no. 99 3729 810 04
• Phoenix, part no. 1543223 SACC-M12MSD-4Q
• When using the device in a hazardous location, use a suitably certified plug.
EtherNet/IP
Pin Assignment
1 + Tx
2 + Rx
3 - Tx
4 - Rx
A0032047
Recommended plug:
• Binder, series 763, part no. 99 3729 810 04
• Phoenix, part no. 1543223 SACC-M12MSD-4Q
• When using the device in a hazardous location, use a suitably certified plug.
Coding Plug/socket
D Socket
Service interface
Order code for "Accessories mounted", option NB: Adapter RJ45 M12 (service interface)
Pin Assignment
1 + Tx
2 + Rx
3 - Tx
4 - Rx
A0032047
Recommended plug:
Endress+Hauser 31
• Binder, series 763, part no. 99 3729 810 04
• Phoenix, part no. 1543223 SACC-M12MSD-4Q
• When using the device in a hazardous location, use a suitably certified plug.
Coding Plug/socket
D Socket
Page 32
Proline Promass F 500

Supply voltage

Order code for "Power supply"
Option D DC24 V ±20%
Option E AC100 to 240 V –15…+10% 50/60 Hz
Option I

Power consumption Transmitter

Max. 10 W (active power)

Current consumption Transmitter

• Max. 400 mA (24 V)
• Max. 200 mA (110 V, 50/60 Hz; 230 V, 50/60 Hz)

Power supply failure

Depending on the device version, the configuration is retained in the device memoryor in the pluggable data memory (HistoROM DAT).
terminal voltage Frequency range
DC24 V ±20%
AC100 to 240 V –15…+10% 50/60 Hz
32 Endress+Hauser
Page 33
Proline Promass F 500
+ AB
61 62 63 64
+ AB
61 62 63 64
4
2
1
4
3
5
6
1
2
4
3
5

Electrical connection Connection of connecting cable: Proline 500 – digital

1 Cable entry for cable on transmitter housing 2 Protective ground (PE) 3 Connecting cable ISEM communication 4 Grounding via ground connection; on device plug versions grounding is through the plug itself 5 Cable entry for cable or connection of device plug on sensor connection housing 6 Protective ground (PE)
A0028198
Depending on the device version of the sensor connection housing, the connecting cable is connected via terminals or device plugs.
Sensor connection housing Order code for "Housing"
Option A: aluminum coated Terminals Terminals
Option B: stainless Terminals Terminals
Option C ultra-compact, hygienic, stainless Device plug Terminals
Option L: cast, stainless Terminals Terminals
Connection to
sensor connection housing
via
Connection to
transmitter housing
via
Pin assignment, device plug
Device plugs are only available for device version, order code for "Housing": Option C ultra-compact, hygienic, stainless For connection to sensor connection housing.
Pin Color
1 Brown + Supply voltage 61
2 White A
3 Blue B
4 Black Supply voltage 63
5 62
1) Cable colors of connecting cable
A connecting cable with a device plug is optionally available.
1)
Coding Plug/socket
A Plug
Assignment Connection
ISEM communication 64
to terminal
Endress+Hauser 33
Page 34
Proline Promass F 500
4
56
7
8
1112 41
42
S1S1S2S2GNDTMTMTTTT
++++
++++
S1S1S2S2GNDTMTMTTTT
2
ER
+
+
ER
ER
ER
5
1
10
9
456789101112 4142
3
4
X3
X4
X3X4
+
TM
2
TM
2
+
TM
2
TM
2
6
6
1 2 3 4 56
Connection of the connecting cable: Proline 500
The connecting cable is connected via terminals.
A0033694
1 Protective ground (PE) 2 Cable entry for connecting cable on transmitter connection housing 3 Connecting cable 4 Cable entry for connecting cable on sensor connection housing 5 Protective ground (PE) 6 Terminals X3, X4: temperature sensor; only for device version with order code for "Test, certificate", option JQ
Connecting the transmitter
• Terminal assignment →  28
• Device plug pin assignment →  30
Connecting the Proline 500 – digital transmitter
A0028200
1 Terminal connection for supply voltage 2 Terminal connection for signal transmission, input/output 3 Terminal connection for signal transmission, input/output 4 Terminal connection for connecting cable between sensor and transmitter 5 Terminal connection for signal transmission, input/output or terminal for network connection (DHCP client)
via service interface (CDI-RJ45); optional: terminal connection for external WLAN antenna
6 Protective ground (PE)
An adapter for RJ45 and the M12 plug is optionally available: Order code for "Accessories", option NB: "Adapter RJ45 M12 (service interface)"
The adapter connects the service interface (CDI-RJ45) to an M12 connector mounted in the cable entry. Therefore the connection to the service interface can be established via an M12 connector without opening the device.
Network connection (DHCP client) via service interface (CDI-RJ45) →  119
34 Endress+Hauser
Page 35
Proline Promass F 500
1
2
3 4
1 2 3 4 56
Connecting the Proline 500 transmitter
A0026781
1 Terminal connection for supply voltage 2 Terminal connection for signal transmission, input/output 3 Terminal connection for signal transmission, input/output or terminal for network connection (DHCP client)
via service interface (CDI-RJ45); optional: terminal connection for external WLAN antenna
4 Protective ground (PE)
An adapter for RJ45 and the M12 plug is optionally available: Order code for "Accessories", option NB: "Adapter RJ45 M12 (service interface)"
The adapter connects the service interface (CDI-RJ45) to an M12 connector mounted in the cable entry. Therefore the connection to the service interface can be established via an M12 connector without opening the device.
Network connection (DHCP client) via service interface (CDI-RJ45) →  119
Connecting in a ring topology
Device versions with EtherNet/IP and PROFINET communication protocols can be integrated into a ring topology. The device is integrated via the terminal connection for signal transmission (output 1) and the connection to the service interface (CDI-RJ45).
Integrating the transmitter into a ring topology:
• EtherNet/IP →  117
• PROFINET →  118
Transmitter: Proline 500 – digital
A0028200
1 Terminal connection for supply voltage 2 Terminal connection for signal transmission, input/output 2 Terminal connection for signal transmission: PROFINET or EtherNet/IP (RJ45 connector) 4 Terminal connection for connecting cable between sensor and transmitter 5 Terminal connection to service interface (CDI-RJ45) 6 Protective ground (PE)
Endress+Hauser 35
Page 36
Proline Promass F 500
1
2
3 4
4
4...20 mA
5
2
1
3
6
2
3
4...20 mA
41
5
Transmitter: Proline 500
1 Terminal connection for supply voltage 2 Terminal connection for signal transmission: PROFINET or EtherNet/IP (RJ45 connector) 3 Terminal connection to service interface (CDI-RJ45) 4 Protective ground (PE)
If the device has additional inputs/outputs, these are routed in parallel via the cable entry for connection to the service interface (CDI-RJ45).
Connection examples
Current output 4 to 20 mA HART
A0026781
A0029055
 2 Connection example for 4 to 20 mA HART current output (active)
1 Automation system with current input (e.g. PLC) 2 Cable shield: the cable shield must be grounded at both ends to comply with EMC requirements; observe cable
specifications →  44 3 Connection for HART operating devices →  113 4 Resistor for HART communication (≥ 250 Ω): observe maximum load →  16 5 Analog display unit: observe maximum load →  16 6 Transmitter
A0028762
 3 Connection example for 4 to 20 mA HART current output (passive)
1 Automation system with current input (e.g. PLC) 2 Power supply 3 Cable shield: the cable shield must be grounded at both ends to comply with EMC requirements; observe cable
specifications →  44
36 Endress+Hauser
4 Analog display unit: observe maximum load →  16 5 Transmitter
Page 37
Proline Promass F 500
2
4...20 mA
4
1
2
3
3
6
5
21 3 4
78
6 6
6
6
5
6
6
5
HART input
A0028763
 4 Connection example for HART input with a common negative (passive)
1 Automation system with HART output (e.g. PLC) 2 Active barrier for power supply (e.g. RN221N) 3 Cable shield: the cable shield must be grounded at both ends to comply with EMC requirements; observe cable
specifications 4 Analog display unit: observe maximum load 5 Pressure transmitter (e.g. Cerabar M, Cerabar S): see requirements 6 Transmitter
PROFIBUS PA
 5 Connection example for PROFIBUS PA
1 Control system (e.g. PLC) 2 PROFIBUS PA segment coupler 3 Cable shield: the cable shield must be grounded at both ends to comply with EMC requirements; observe cable
4 T-box 5 Measuring device 6 Local grounding 7 Bus terminator 8 Potential matching line
specifications
A0028768
Endress+Hauser 37
Page 38
Proline Promass F 500
2
1
A
B
3
4
4
A
B
A
B
1
2
4
3
5
5
PROFIBUS DP
A0028765
 6 Connection example for PROFIBUS DP, non-hazardous area and Zone 2/Div. 2
1 Control system (e.g. PLC) 2 Cable shield: the cable shield must be grounded at both ends to comply with EMC requirements; observe cable
specifications 3 Distribution box 4 Transmitter
If baud rates > 1.5 MBaud an EMC cable entry must be used and the cable shield must continue as far as the terminal wherever possible.
EtherNet/IP
 7 Connection example for EtherNet/IP
1 Control system (e.g. PLC) 2 Ethernet switch 3 Observe cable specifications 4 Device plug 5 Transmitter
A0028767
38 Endress+Hauser
Page 39
Proline Promass F 500
1 2 4
5
5
3
1 2 4
5
5
3
1 2 4
5
5
3
EtherNet/IP: DLR (Device Level Ring)
A0027544
1 Control system (e.g. PLC) 2 Ethernet switch 3 Observe cable specifications →  44 4 Connecting cable between the two transmitters 5 Transmitter
PROFINET
 8 Connection example for PROFINET
1 Control system (e.g. PLC) 2 Ethernet switch 3 Observe cable specifications 4 Device plug 5 Transmitter
PROFINET: MRP (Media Redundancy Protocol)
1 Control system (e.g. PLC) 2 Ethernet switch 3 Observe cable specifications →  44 4 Connecting cable between the two transmitters 5 Transmitter
A0016805
A0027544
Endress+Hauser 39
Page 40
FOUNDATION Fieldbus
21 3 4
78
6 6
6
6
5
6
6
5
2
1
A
B
3
4
4
A
B
A
B
Proline Promass F 500
A0028768
 9 Connection example for FOUNDATION Fieldbus
1 Control system (e.g. PLC) 2 Power Conditioner (FOUNDATION Fieldbus) 3 Cable shield: the cable shield must be grounded at both ends to comply with EMC requirements; observe cable
specifications 4 T-box 5 Measuring device 6 Local grounding 7 Bus terminator 8 Potential matching line
Modbus RS485
A0028765
 10 Connection example for Modbus RS485, non-hazardous area and Zone 2; Class I, Division 2
1 Control system (e.g. PLC) 2 Cable shield: the cable shield must be grounded at both ends to comply with EMC requirements; observe cable
specifications 3 Distribution box 4 Transmitter
40 Endress+Hauser
Page 41
Proline Promass F 500
4...20 mA
2
1
3
2
4...20 mA
3
1
4
1
2
3
12345
Current output 4-20 mA
A0028758
 11 Connection example for 4-20 mA current output (active)
1 Automation system with current input (e.g. PLC) 2 Analog display unit: observe maximum load 3 Transmitter
 12 Connection example for 4-20 mA current output (passive)
1 Automation system with current input (e.g. PLC) 2 Active barrier for power supply (e.g. RN221N) 3 Analog display unit: observe maximum load 4 Transmitter
Pulse/frequency output
 13 Connection example for pulse/frequency output (passive)
1 Automation system with pulse/frequency input (e.g. PLC) 2 Power supply 3 Transmitter: Observe input values →  17
A0028759
A0028761
Endress+Hauser 41
Page 42
Switch output
1
2
3
1
2
3
4
1
3
2
4
5
 14 Connection example for switch output (passive)
1 Automation system with switch input (e.g. PLC) 2 Power supply 3 Transmitter: Observe input values →  17
Double pulse output
Proline Promass F 500
A0028760
A0029280
 15 Connection example for double pulse output (active)
1 Automation system with double pulse input (e.g. PLC) 2 Transmitter: Observe input values →  19 3 Double pulse output 4 Double pulse output (slave), phase-shifted
A0029279
 16 Connection example for double pulse output (passive)
1 Automation system with double pulse input (e.g. PLC) 2 Power supply 3 Transmitter: Observe input values →  19 4 Double pulse output 5 Double pulse output (slave), phase-shifted
42 Endress+Hauser
Page 43
Proline Promass F 500
1
2
3
31
4
2
1
2
3
Relay output
A0028760
 17 Connection example for relay output (passive)
1 Automation system with relay input (e.g. PLC) 2 Power supply 3 Transmitter: Observe input values →  19
Current input
 18 Connection example for 4 to 20 mA current input
1 Power supply 2 Terminal box 3 External measuring device (for reading in pressure or temperature, for instance) 4 Transmitter
Status input
 19 Connection example for status input
1 Automation system with status output (e.g. PLC) 2 Power supply 3 Transmitter
A0028915
A0028764

Potential equalization Requirements

No special measures for potential equalization are required.
Endress+Hauser 43
Page 44
Please consider the following to ensure correct measurement:
• Same electrical potential for the fluid and sensor
• Company-internal grounding concepts
Proline Promass F 500

terminals

Spring-loaded terminals: Suitable for strands and strands with ferrules. Conductor cross-section 0.2 to 2.5 mm2 (24 to 12 AWG).

Cable entries

• Cable gland: M20 × 1.5 with cable ⌀ 6 to 12 mm (0.24 to 0.47 in)
• Thread for cable entry:
– NPT ½" – G ½" – M20
• Device plug for digital communication: M12
Only available for certain device versions →  29.
• Device plug for connecting cable: M12
A device plug is always used for the device version with the order code for "Sensor connection housing", option C "Ultra-compact, hygienic, stainless".

Cable specification Permitted temperature range

• The installation guidelines that apply in the country of installation must be observed.
• The cables must be suitable for the minimum and maximum temperatures to be expected.
Power supply cable
Standard installation cable is sufficient.
Protective ground cable
Cable ≥2.08 mm2 (14 AWG)
The grounding impedance must be less than 1 Ω.
Signal cable
Current output 4 to 20 mA HART
A shielded cable is recommended. Observe grounding concept of the plant.
PROFIBUS PA
Twisted, shielded two-wire cable. Cable type A is recommended .
For further information on planning and installing PROFIBUS networks see:
• Operating Instructions "PROFIBUS DP/PA: Guidelines for planning and commissioning" (BA00034S)
• PNO Directive 2.092 "PROFIBUS PA User and Installation Guideline"
• IEC 61158-2 (MBP)
PROFIBUS DP
The IEC 61158 standard specifies two types of cable (A and B) for the bus line which can be used for every transmission rate. Cable type A is recommended.
Cable type A
Characteristic impedance 135 to 165 Ω at a measuring frequency of 3 to 20 MHz
Cable capacitance < 30 pF/m
Wire cross-section
Cable type Twisted pairs
Loop resistance ≤110 Ω/km
> 0.34 mm2 (22 AWG)
44 Endress+Hauser
Page 45
Proline Promass F 500
Signal damping Max. 9 dB over the entire length of the cable cross-section
Shield Copper braided shielding or braided shielding with foil shield. When grounding
the cable shield, observe the grounding concept of the plant.
For further information on planning and installing PROFIBUS networks see:
• Operating Instructions "PROFIBUS DP/PA: Guidelines for planning and commissioning" (BA00034S)
• PNO Directive 2.092 "PROFIBUS PA User and Installation Guideline"
• IEC 61158-2 (MBP)
EtherNet/IP
The standard ANSI/TIA/EIA-568-B.2 Annex specifies CAT 5 as the minimum category for a cable used for EtherNet/IP. CAT 5e and CAT 6 are recommended.
For more information on planning and installing EtherNet/IP networks, please refer to the "Media Planning and Installation Manual. EtherNet/IP" of ODVA Organization
PROFINET
Standard IEC 61156-6 specifies CAT 5 as the minimum category for a cable used for PROFINET. CAT 5e and CAT 6 are recommended.
For more information on planning and installing PROFINET networks, see: "PROFINET Cabling and Interconnection Technology", Guideline for PROFINET
FOUNDATION Fieldbus
Twisted, shielded two-wire cable.
For further information on planning and installing FOUNDATION Fieldbus networks see:
• Operating Instructions for "FOUNDATION Fieldbus Overview" (BA00013S)
• FOUNDATION Fieldbus Guideline
• IEC 61158-2 (MBP)
Modbus RS485
The EIA/TIA-485 standard specifies two types of cable (A and B) for the bus line which can be used for every transmission rate. Cable type A is recommended.
Cable type A
Characteristic impedance 135 to 165 Ω at a measuring frequency of 3 to 20 MHz
Cable capacitance < 30 pF/m
Wire cross-section
Cable type Twisted pairs
Loop resistance ≤110 Ω/km
Signal damping Max. 9 dB over the entire length of the cable cross-section
Shield Copper braided shielding or braided shielding with foil shield. When grounding
> 0.34 mm2 (22 AWG)
the cable shield, observe the grounding concept of the plant.
Current output 0/4 to 20 mA
Standard installation cable is sufficient.
Pulse/frequency/switch output
Standard installation cable is sufficient.
Double pulse output
Standard installation cable is sufficient.
Endress+Hauser 45
Page 46
Relay output
2
3
C
1
3
B
1
3
A
1
3
A
1
3
B
2
3
C
4
5
6
Standard installation cable is sufficient.
Current input 0/4 to 20 mA
Standard installation cable is sufficient.
Status input
Standard installation cable is sufficient.
Choice of connecting cable between the transmitter and sensor
Depends on the type of transmitter and the installation zones
Proline Promass F 500
1 Proline 500 digital transmitter 2 Proline 500 transmitter 3 Sensor Promass 4 Non-hazardous area 5 Hazardous area: Zone 2; Class I, Division 2 6 Hazardous area: Zone 1; Class I, Division 1 A Standard cable to 500 digital transmitter →  46
Transmitter installed in the non-hazardous area or hazardous area: Zone 2; Class I, Division 2 / sensor installed in the hazardous area: Zone 2; Class I, Division 2
B Standard cable to 500 digital transmitter →  47
Transmitter installed in the hazardous area: Zone 2; Class I, Division 2 / sensor installed in the hazardous area: Zone 1; Class I, Division 1
C Signal cable to 500 transmitter →  49
Transmitter and sensor installed in the hazardous area: Zone 2; Class I, Division 2 oder Zone 1; Class I, Division 1
A: Connecting cable between sensor and transmitter: Proline 500 – digital
Standard cable
A standard cable with the following specifications can be used as the connecting cable.
Design 4 cores (2 pairs); uninsulated stranded CU wires; pair-stranded with common
Shielding Tin-plated copper-braid, optical cover ≥ 85 %
Loop resistance Power supply line (+, –): maximum 10 Ω
Cable length Maximum 300 m (1 000 ft), see the following table.
shield
46 Endress+Hauser
A0032476
Page 47
Proline Promass F 500
Cross-section Cable length [max.]
0.34 mm2 (AWG 22) 80 m (270 ft)
0.50 mm2 (AWG 20) 120 m (400 ft)
0.75 mm2 (AWG 18) 180 m (600 ft)
1.00 mm2 (AWG 17) 240 m (800 ft)
1.50 mm2 (AWG 15) 300 m (1 000 ft)
Optionally available connecting cable
Design
Flame resistance According to DIN EN 60332-1-2
Oil-resistance According to DIN EN 60811-2-1
Shielding Tin-plated copper-braid, optical cover ≥ 85 %
Operating temperature When mounted in a fixed position: –50 to +105 °C (–58 to +221 °F); when cable
Available cable length Fixed: 20 m (65 ft); variable: up to maximum 50 m (165 ft)
1) UV radiation can impair the cable outer sheath. Protect the cable from direct sunshine where possible.
2 × 2 × 0.34 mm2 (AWG 22) PVC cable uninsulated stranded CU wires; pair-stranded)
can move freely: –25 to +105 °C (–13 to +221 °F)
1)
with common shield (2 pairs,
B: Connecting cable between sensor and transmitter: Proline 500 - digital
Standard cable
A standard cable with the following specifications can be used as the connecting cable.
Design 4, 6, 8 cores (2, 3, 4 pairs); uninsulated stranded CU wires; pair-stranded with
common shield
Shielding Tin-plated copper-braid, optical cover ≥ 85 %
Capacitance C Maximum 760 nF IIC, maximum 4.2 µF IIB
Inductance L Maximum 26 µH IIC, maximum 104 µH IIB
Inductance/resistance ratio (L/R)
Loop resistance Power supply line (+, –): maximum 5 Ω
Cable length Maximum 150 m (500 ft), see the following table.
Maximum 8.9 µH/Ω IIC, maximum 35.6 µH/Ω IIB (e.g. in accordance with IEC 60079-25)
Endress+Hauser 47
Page 48
Cross-section Cable length [max.] Termination
A B
GY
BN WT YE GN
A B
GY
BN
YE GN
WT
GY PK
A B
YE GN
GY
BN
RD BU
WT
GY PK
2 x 2 x 0.50 mm
2
50 m (165 ft) 2 x 2 x 0.50 mm2 (AWG 20)
(AWG 20)
Proline Promass F 500
3 x 2 x 0.50 mm (AWG 20)
4 x 2 x 0.50 mm (AWG 20)
• +, – = 0.5 mm
• A, B = 0.5 mm
2
100 m (330 ft) 3 x 2 x 0.50 mm2 (AWG 20)
• +, – = 1.0 mm
• A, B = 0.5 mm
2
150 m (500 ft) 4 x 2 x 0.50 mm2 (AWG 20)
2
2
2
2
• +, – = 1.5 mm
• A, B = 0.5 mm
2
2
Optionally available connecting cable
Connecting cable for Zone 1; Class I, Division 1
Standard cable
2 × 2 × 0.5 mm2 (AWG 20) PVC cable stranded)
Flame resistance According to DIN EN 60332-1-2
Oil-resistance According to DIN EN 60811-2-1
Shielding Tin-plated copper-braid, optical cover ≥ 85 %
Operating temperature When mounted in a fixed position: –50 to +105 °C (–58 to +221 °F); when cable
can move freely: –25 to +105 °C (–13 to +221 °F)
Available cable length Fixed: 20 m (65 ft); variable: up to maximum 50 m (165 ft)
1) UV radiation can impair the cable outer sheath. Protect the cable from direct sunshine where possible.
1)
with common shield (2 pairs, pair-
48 Endress+Hauser
Page 49
Proline Promass F 500
C: Connecting cable between sensor and transmitter: Proline 500
Standard cable 6 × 0.38 mm2 PVC cable
cores
Conductor resistance ≤50 Ω/km (0.015 Ω/ft)
Capacitance: core/shield ≤420 pF/m (128 pF/ft)
Cable length (max.) 20 m (65 ft)
Cable lengths (available for order)
Operating temperature Depends on the device version and how the cable is installed:
1) UV radiation can impair the cable outer sheath. Protect the cable from direct sunshine where possible.
5 m (15 ft), 10 m (32 ft), 20 m (65 ft)
• Standard version: – Cable - fixed installation: –40 to +105 °C (–40 to +221 °F) – Cable - movable: –25 to +105 °C (–13 to +221 °F)
• Order code for "Test, certificate", option JP: – Cable - fixed installation: –50 to +105 °C (–58 to +221 °F) – Cable - movable: –25 to +105 °C (–13 to +221 °F)
• Order code for "Test, certificate", option JQ: – Cable - fixed installation: –60 to +105 °C (–76 to +221 °F) – Cable - movable: –25 to +105 °C (–13 to +221 °F)
1)
with common shield and individually shielded

Performance characteristics

Reference operating conditions

Maximum measured error

• Error limits based on ISO 11631
• Water with +15 to +45 °C (+59 to +113 °F) at2 to 6 bar (29 to 87 psi)
• Specifications as per calibration protocol
• Accuracy based on accredited calibration rigs that are traced to ISO 17025. To obtain measured errors, use the Applicator sizing tool →  134
o.r. = of reading; 1 g/cm³ = 1 kg/l; T = medium temperature
Base accuracy
Design fundamentals →  53
Mass flow and volume flow (liquids)
±0.05 % o.r. (PremiumCal; order code for "Calibration flow", option D, for mass flow) ±0.10 % o.r.
Mass flow (cryogenic liquids)
Order code for "Measuring tube material", option LA
±0.35 % o.r.
Mass flow (gases)
±0.35 % o.r.
Endress+Hauser 49
Page 50
Density (liquids)
Proline Promass F 500
Under
reference operating conditions
[g/cm³] [g/cm³] [g/cm³]
±0.0005 ±0.01 ±0.001
1) Valid over the entire temperature and density range
2) Valid range for special density calibration: 0 to 2 g/cm³, +5 to +80 °C (+41 to +176 °F)
3) Order code for "Application package", option EE "Special density" only in combination with the order code for "Measuring tube mat., wetted surface", option BB, BF, HA, SA
Standard density
calibration
1)
Wide-range
Density specification
2) 3)
Density (cryogenic liquids)
Order code for "Measuring tube material", option LA
±0.05 g/cm³
Temperature
±0.5 °C ± 0.005 · T °C (±0.9 °F ± 0.003 · (T – 32) °F)
Zero point stability
DN Zero point stability
[mm] [in] [kg/h] [lb/min]
8 ³⁄₈ 0.030 0.001
15 ½ 0.200 0.007
25 1 0.540 0.019
40 2.25 0.083
50 2 3.50 0.129
80 3 9.0 0.330
100 4 14.0 0.514
150 6 32.0 1.17
250 10 88.0 3.23
High-temperature version: order code for "Measuring tube material", option TT, TU
DN Zero point stability
[mm] [in] [kg/h] [lb/min]
25 1 1.80 0.0661
50 2 7.00 0.2572
80 3 18.0 0.6610
For devices with low-temperature version, order code for "Measuring tube mat., wetted surface", option LA, please note the following:
NOTICE
Zero point confirmation and zero point adjustment are difficult to carry out in the field due to the vaporization of the cryogenic liquid.
As a general rule, the factory-set zero point should not be changed. Please ensure that the
medium is in the liquid phase if a zero point adjustment is to be carried out.
Flow values
Flow values as turndown parameter depending on nominal diameter.
50 Endress+Hauser
Page 51
Proline Promass F 500
SI units
DN 1:1 1:10 1:20 1:50 1:100 1:500
[mm]
8 2 000 200 100 40 20 4
15 6 500 650 325 130 65 13
25 18 000 1 800 900 360 180 36
40 45 000 4 500 2 250 900 450 90
50 70 000 7 000 3 500 1 400 700 140
80 180 000 18 000 9 000 3 600 1 800 360
100 350 000 35 000 17 500 7 000 3 500 700
150 800 000 80 000 40 000 16 000 8 000 1 600
250 2 200 000 220 000 110 000 44 000 22 000 4 400
[kg/h] [kg/h] [kg/h] [kg/h] [kg/h] [kg/h]
US units
DN 1:1 1:10 1:20 1:50 1:100 1:500
[inch] [lb/min] [lb/min] [lb/min] [lb/min] [lb/min] [lb/min]
³⁄₈ 73.50 7.350 3.675 1.470 0.735 0.147
½ 238.9 23.89 11.95 4.778 2.389 0.478
1 661.5 66.15 33.08 13.23 6.615 1.323
1 654 165.4 82.70 33.08 16.54 3.308
2 2 573 257.3 128.7 51.46 25.73 5.146
3 6 615 661.5 330.8 132.3 66.15 13.23
4 12 860 1 286 643.0 257.2 128.6 25.72
6 29 400 2 940 1 470 588 294 58.80
10 80 850 8 085 4 043 1 617 808.5 161.7
Accuracy of outputs
The outputs have the following base accuracy specifications.
Current output
Accuracy ±5 µA
Pulse/frequency output
o.r. = of reading
Accuracy Max. ±50 ppm o.r. (over the entire ambient temperature range)

Repeatability

o.r. = of reading; 1 g/cm3 = 1 kg/l; T = medium temperature
Base repeatability
Design fundamentals →  53
Mass flow and volume flow (liquids)
±0.025 % o.r. (PremiumCal, for mass flow) ±0.05 % o.r.
Endress+Hauser 51
Page 52
Mass flow (cryogenic liquids)
Order code for "Measuring tube material", option LA ±0.175 % % o.r.
Mass flow (gases)
±0.25 % o.r.
Density (liquids)
±0.00025 g/cm
3
Density (cryogenic liquids)
Order code for "Measuring tube material", option LA ±0.025 g/cm
3
Temperature
±0.25 °C ± 0.0025 · T °C (±0.45 °F ± 0.0015 · (T–32) °F)
Proline Promass F 500

Response time

Influence of ambient temperature

Influence of medium temperature

The response time depends on the configuration (damping).
Current output
Temperature coefficient Max. 1 μA/°C
Pulse/frequency output
Temperature coefficient No additional effect. Included in accuracy.
Mass flow and volume flow
o.f.s. = of full scale value
When there is a difference between the temperature for zero point adjustment and the process temperature, the additional measured error of the sensor is typically ±0.0002 % o.f.s./°C (±0.0001 % o. f.s./°F).
The effect is reduced if zero point adjustment is performed at process temperature.
Density
When there is a difference between the density calibration temperature and the process temperature, the typical measured error of the sensor is ±0.00005 g/cm3 /°C (±0.000025 g/cm3 /°F). Field density calibration is possible.
Wide-range density specification (special density calibration)
If the process temperature is outside the valid range (→  49) the measured error is ±0.00005 g/cm3 /°C (±0.000025 g/cm3 /°F)
52 Endress+Hauser
Page 53
Proline Promass F 500
3
[°C]
0- 08 80 160 240
[° ]F
- 05 0
100
200
300
0
2
4
6
8
10
2
1
12
14
16
18
320 400 480 560
640
A0027453
1 Field density calibration, for example at +20 °C (+68 °F) 2 Special density calibration
Temperature
±0.005 · T °C (± 0.005 · (T – 32) °F)

Influence of medium pressure

The table below shows the effect on accuracy of mass flow due to a difference between calibration pressure and process pressure.
o.r. = of reading
It is possible to compensate for the effect by:
• Reading in the current pressure measured value via the current input.
• Specifying a fixed value for the pressure in the device parameters. Operating Instructions→  135.
DN [% o.r./bar] [% o.r./psi]
[mm] [in]
8 3/8 no influence
15 ½ no influence
25 1 no influence
40 –0.003 –0.0002
50 2 –0.008 –0.0006
80 3 –0.009 –0.0006
100 4 –0.007 –0.0005

Design fundamentals

150 6 –0.009 –0.0006
250 10 –0.009 –0.0006
o.r. = of reading, o.f.s. = of full scale value
BaseAccu = base accuracy in % o.r., BaseRepeat = base repeatability in % o.r.
MeasValue = measured value; ZeroPoint = zero point stability
Endress+Hauser 53
Page 54
Calculation of the maximum measured error as a function of the flow rate
ZeroPoint BaseAccu
100
³
± BaseAccu
ZeroPoint BaseAccu
100
<
ZeroPoint
MeasValue
100
±
½ ZeroPoint
BaseRepeat
100
³
½ ZeroPoint
BaseRepeat
100
<
ZeroPoint
MeasValue
100
½ ±
E [%]
0 10 20 30 40 50 60
0
0.5
1.0
1.5
2.0
2.5
70 80 90 100
Q [%]
Flow rate Maximum measured error in % o.r.
Proline Promass F 500
A0021332
A0021333
Calculation of the maximum repeatability as a function of the flow rate
Flow rate Maximum repeatability in % o.r.
A0021335
A0021336
Example for maximum measured error
A0021339
A0021334
A0021340
A0021337
E Maximum measured error in % o.r. (example with PremiumCal) Q Flow rate in % of maximum full scale value

Installation

No special measures such as supports etc. are necessary. External forces are absorbed by the construction of the device.

Mounting location

54 Endress+Hauser
To prevent measuring errors arising from accumulation of gas bubbles in the measuring tube, avoid the following mounting locations in the pipe:
• Highest point of a pipeline.
• Directly upstream of a free pipe outlet in a down pipe.
A0028808
A0028772
Page 55
Proline Promass F 500
1
2
3
4
5
Installation in down pipes
However, the following installation suggestion allows for installation in an open vertical pipeline. Pipe restrictions or the use of an orifice with a smaller cross-section than the nominal diameter prevent the sensor running empty while measurement is in progress.
A0028773
 20 Installation in a down pipe (e.g. for batching applications)
1 Supply tank 2 Sensor 3 Orifice plate, pipe restriction 4 Valve 5 Batching tank

Orientation

DN Ø orifice plate, pipe restriction
[mm] [in] [mm] [in]
8 ³⁄₈ 6 0.24
15 ½ 10 0.40
25 1 14 0.55
40 22 0.87
50 2 28 1.10
80 3 50 1.97
100 4 65 2.60
150 6 90 3.54
250 10 150 5.91
The direction of the arrow on the sensor nameplate helps you to install the sensor according to the flow direction (direction of medium flow through the piping).
Orientation Recommendation
A Vertical orientation
B Horizontal orientation, transmitter at
top
A0015591
1)
Exceptions:
→  21,  56
A0015589
Endress+Hauser 55
Page 56
Proline Promass F 500
1 2
Orientation Recommendation
C Horizontal orientation, transmitter at
bottom
→  21,  56
A0015590
D Horizontal orientation, transmitter at
side
A0015592
1) Applications with low process temperatures may decrease the ambient temperature. To maintain the minimum ambient temperature for the transmitter, this orientation is recommended.
2) Applications with high process temperatures may increase the ambient temperature. To maintain the maximum ambient temperature for the transmitter, this orientation is recommended.
If a sensor is installed horizontally with a curved measuring tube, match the position of the sensor to the fluid properties.
2)
Exceptions:

Inlet and outlet runs

A0028774
 21 Orientation of sensor with curved measuring tube
1 Avoid this orientation for fluids with entrained solids: Risk of solids accumulating. 2 Avoid this orientation for outgassing fluids: Risk of gas accumulating.
No special precautions need to be taken for fittings which create turbulence, such as valves, elbows or T-pieces, as long as no cavitation occurs →  69.
56 Endress+Hauser
Page 57
Proline Promass F 500
ø 20…70
( 0.79…2.75)ø
~102 (~ 4.0)
4 x
SW 10
3 x
TX 25
149 (5.85)
211 (8.31)
=
5.8 (0.23)
17 (0.67) =
L
5.8 (0.23)

Mounting the transmitter housing

Proline 500 – digital transmitter
Post mounting
A0029051
 22 Engineering unit mm (in)
Wall mounting
 23 Engineering unit mm (in)
L Depends on order code for "Transmitter housing"
Order code for "Transmitter housing"
• Option A, aluminum coated: L =14 mm (0.55 in)
• Option D, polycarbonate: L = 13 mm (0.51 in)
A0029054
Endress+Hauser 57
Page 58
Proline Promass F 500
1
4 x
SW 13
! (! 0.79 to 2.75)20 70
20 (0.79)
100 (3.94)
100 (3.94)
! 8.6 (0.39)
! 18 (0.71)
! 10 (0.39)
Proline 500 transmitter
Post mounting
WARNING
L
Order code for "Transmitter housing", option L "Cast, stainless": cast transmitters are very heavy.
They are unstable if they are not mounted on a secure, fixed post.
Only mount the transmitter on a secure, fixed post on a stable surface.

Special mounting instructions

 24 Engineering unit mm (in)
Wall mounting
 25 Engineering unit mm (in)
Rupture disk
Information that is relevant to the process: →  69.
A0029057
A0029068
58 Endress+Hauser
Page 59
Proline Promass F 500
RUPTURE DISK
i
2
3
1
DN 8 ( ")...150 (6")!⁄# DN 250 (10")
2
3
WARNING
L
Danger from medium escaping!
Medium escaping under pressure can cause injury or material damage.
Take precautions to prevent danger to persons and damage if the rupture disk is actuated.
Observe information on the rupture disk sticker.
Make sure that the function and operation of the rupture disk is not impeded through the
installation of the device. Do not use a heating jacket.
Do not remove or damage the rupture disk.
The position of the rupture disk is indicated on a sticker beside it.
The transportation guard must be removed.
The existing connecting nozzles are not intended for the purpose of rinsing or pressure monitoring, but instead serve as the mounting location for the rupture disk.
In the event of a failure of the rupture disk, a discharge device can be screwed onto the internal thread of the rupture disk in order to drain off any escaping medium.
1 Rupture disk label 2 Rupture disk with 1/2" NPT internal thread with 1" width across flat 3 Transport protection
For information on the dimensions: see the "Mechanical construction" section (accessories)
Zero point adjustment
All measuring devices are calibrated in accordance with state-of-the-art technology. Calibration takes place under reference conditions→  49. Therefore, a zero point adjustment in the field is generally not required.
Experience shows that zero point adjustment is advisable only in special cases:
• To achieve maximum measuring accuracy even with low flow rates.
• Under extreme process or operating conditions (e.g. very high process temperatures or very high­viscosity fluids).
A0028903
Endress+Hauser 59
Page 60

Environment

Proline Promass F 500

Ambient temperature range

Storage temperature

Climate class

Degree of protection

Measuring device • –40 to +60 °C (–40 to +140 °F)
• Order code for "Test, certificate", option JP: –50 to +60 °C (–58 to +140 °F)
• Order code for "Test, certificate", option JQ: – Sensor:–60 to +60 °C (–76 to +140 °F) – Transmitter: –50 to +60 °C (–58 to +140 °F)
Readability of the local display
–20 to +60 °C (–4 to +140 °F) The readability of the display may be impaired at temperatures outside the temperature range.
Dependency of ambient temperature on medium temperature→  61
If operating outdoors:
Avoid direct sunlight, particularly in warm climatic regions.
You can order a weather protection cover from Endress+Hauser. →  132.
–50 to +80 °C (–58 to +176 °F)
DIN EN 60068-2-38 (test Z/AD)
Transmitter
• As standard: IP66/67, type 4X enclosure
• When housing is open: IP20, type 1 enclosure
• Display module: IP20, type 1 enclosure
Sensor
• As standard: IP66/67, type 4X enclosure
• With the order code for "Sensor options", option CM: IP69 can also be ordered
External WLAN antenna
IP67

Vibration resistance

Device version: Order code for "Sensor connection housing", option: A "Aluminum, coated", B "Stainless" or C "Ultra-compact hygienic, stainless"
• Vibration, sinusoidal according to IEC 60068-2-6 – 2 to 8.4 Hz, 7.5 mm peak – 8.4 to 2 000 Hz, 2 g peak
• Vibration broad-band random, according to IEC 60068-2-64 – 10 to 200 Hz, 0.01 g2/Hz – 200 to 2 000 Hz, 0.003 g2/Hz – Total: 2.70 g rms
Device version: Order code for "Sensor connection housing", option: L "Cast, stainless"
For medium temperature range: standard version
• Vibration, sinusoidal according to IEC 60068-2-6 – 2 to 8.4 Hz, 7.5 mm peak – 8.4 to 2 000 Hz, 2 g peak
• Vibration broad-band random, according to IEC 60068-2-64 – 10 to 200 Hz, 0.01 g2/Hz – 200 to 2 000 Hz, 0.003 g2/Hz – Total: 2.70 g rms
60 Endress+Hauser
Page 61
Proline Promass F 500
Device version: Order code for "Sensor connection housing", option: L "Cast, stainless"
For medium temperature range: extended temperature version, high-temperature version and low­temperature version
• Vibration, sinusoidal according to IEC 60068-2-6 – 2 to 8.4 Hz, 3.5 mm peak – 8.4 to 2 000 Hz, 1 g peak
• Vibration broad-band random, according to IEC 60068-2-64 – 10 to 200 Hz, 0.003 g2/Hz – 200 to 2 000 Hz, 0.001 g2/Hz – Total: 1.54 g rms

Shock resistance

Shock resistance

Interior cleaning

Electromagnetic compatibility (EMC)

Shock, half-sine according to IEC 60068-2-27 6 ms 50 g
Shock due to rough handling following IEC 60068-2-31
• Cleaning in place (CIP)
• Sterilization in place (SIP)
Options
• Oil- and grease-free version for wetted parts, without declaration Order code for "Service", option HA
• Oil- and grease-free version for wetted parts as per IEC/TR 60877-2.0 and BOC 50000810-4, with declaration Order code for "Service", option HB
• As per IEC/EN 61326 and NAMUR Recommendation 21 (NE 21)
• Device version with PROFIBUS DP: Complies with emission limits for industry as per EN 50170 Volume 2, IEC 61784
The following applies for PROFIBUS DP: If baud rates > 1.5 MBaud, an EMC cable entry must be used and the cable shield must continue as far as the terminal wherever possible.
Details are provided in the Declaration of Conformity.

Process

Medium temperature range

Standard version –50 to +150 °C (–58 to +302 °F) Order code for "Measuring tube
mat., wetted surface", option HA, SA, SB, SC
Extended temperature version –50 to +240 °C (–58 to +464 °F) Order code for "Measuring tube
mat., wetted surface", option SD, SE, SF, TH
High-temperature version –50 to +350 °C (–58 to +662 °F) For nominal diameters DN 25 (1"),
DN 50 (2") and DN 80 (3") Order code for "Measuring tube mat., wetted surface", option TT, TU
Low-temperature version –196 to +150 °C (–320 to +302 °F)
NOTICE
Material fatigue due to excessive temperature difference!
Maximum temperature
difference of media used: 300 K
Order code for "Measuring tube mat., wetted surface", option LA
Endress+Hauser 61
Page 62
Proline Promass F 500
T
a
T
m
B
A
Dependency of ambient temperature on medium temperature
 26 Exemplary representation, values in the table below.
TaAmbient temperature TmMedium temperature A Maximum permitted medium temperature Tm at T
require a reduced ambient temperature T
a
B Maximum permitted ambient temperature Ta for the maximum specified medium temperature Tm of the
sensor
= 60 °C (140 °F); higher medium temperatures T
a max
A0031121
m
Values for devices used in the hazardous area: Separate Ex documentation (XA) for the device →  136.
Not insulated Insulated
A B A B
1)
Version
Standard version 60 °C
Extended temperature version 60 °C
High-temperature version 60 °C
1) The values apply for Promass F 500 - digital and Promass F 500.

Density

T
a
(140 °F)
T
m
130 °C
(266 °F)
55 °C
(131 °F)
240 °C
(140 °F)
(140 °F)
(464 °F)
240 °C
(464 °F)
50 °C
(122 °F)
0 to 5 000 kg/m3 (0 to 312 lb/cf)
T
a
T
m
150 °C
(302 °F)
T
60 °C
(140 °F)
60 °C
(140 °F)
350 °C
(662 °F)
60 °C
(140 °F)
a
T
m
90 °C
(194 °F)
150 °C
(302 °F)
210 °C
(410 °F)
T
a
45 °C
(113 °F)
50 °C
(122 °F)
50 °C
(122 °F)
T
m
150 °C
(302 °F)
240 °C
(464 °F)
350 °C
(662 °F)
62 Endress+Hauser
Page 63
Proline Promass F 500
PN100
PN63
PN40
PN16
0
10
20
30
40
50
60
70
80
90
100
0
200
400
600
800
[bar][psi]
-320
64080 160
240
320 400 480 560
[°C]
[°F]
35030025020015010050
0
-200

Pressure-temperature curves

The following pressure/temperature diagrams apply to all pressure-bearing parts of the device and not just the process connection. The diagrams show the maximum permissible medium pressure depending on the specific medium temperature.
The medium temperature range of the device depends on the device version →  61.
The values in the pressure-temperature curves only apply for the medium temperature range of the specific device version.
The extended ambient temperature range for device versions with the order code for "Test, certificate", option JQ (sensor –60 to +60 °C (–76 to +140 °F)) does not affect the values in the pressure-temperature curves.
The value for the minimum ambient temperature –60 °C (–76 °F) corresponds to the value for the minimum medium temperature of the standard version indicated in the particular pressure­temperature curve –50 °C (–58 °F).
• Pressure-temperature curves with temperature range +151 to +240 °C (+304 to +464 °F) exclusively for extended temperature version of measuring devices.
• Pressure-temperature curves with temperature range +241 to +350 °C (+466 to +662 °F) exclusively for high temperature version of measuring devices.
• Pressure-temperature curves with temperature range –196 to +150 °C (–320 to +302 °F) exclusively for low temperature version of measuring devices.
Flange according to EN 1092-1 (DIN 2501)
A0032852-EN
 27 With flange material 1.4404 (F316/F316L), Alloy C22
Endress+Hauser 63
Page 64
Flange according to ASME B16.5
Class 600
Class 300
Class 150
0
10
20
30
40
50
60
70
80
90
100
0
200
400
600
800
[bar][psi]
-320
64080 160
240
320 400 480 560
[°C]
[°F]
35030025020015010050
0
-200
Class 600
Class 300
Class 150
-80
6400 80 160
240
320 400 480 560
0
10
20
30
40
50
60
70
80
90
100
0
200
400
600
800
[°C]
[°F]
[bar][psi]
35030025020015010050
0
-50
 28 With flange material 1.4404 (F316/F316L)
Proline Promass F 500
A0032853-EN
 29 With flange material Alloy C22
64 Endress+Hauser
A0028780-EN
Page 65
Proline Promass F 500
63K
40K
20K
10K
0
10
20
30
40
50
60
70
0
200
400
600
800
[bar][psi]
-320
64080 160
240
320 400 480 560
[°C]
[°F]
35030025020015010050
0
-200
DN8…40
≥ DN50
-80
0 80 160
240
320 400 480
0
10
20
30
40
0
[°C]
[°F]
[bar][psi]
25020015010050
0
-50
50
PN40
-80
0 80 160
240
320 400 480
0
10
20
30
40
0
[°C]
[°F]
[bar][psi]
25020015010050
0
-50
50
Flange JIS B2220
A0032851-EN
 30 With flange material 1.4404 (F316/F316L), Alloy C22
Flange DIN 11864-2 Form A
A0028782-EN
 31 With connection material 1.4404 (316/316L)
Lap joint flange according to EN 1092-1 (DIN 2501)
A0028784-EN
 32 With flange material 1.4301 (F304); wetted parts Alloy C22
Endress+Hauser 65
Page 66
Lap joint flange according to ASME B16.5
Class 150
Class 600
Class 300
-80
0 80 160
240
320 400 480
0
10
20
30
40
50
60
70
80
90
100
0
200
400
600
800
[°C]
[°F]
[bar][psi]
25020015010050
0
-50
20K
-80
0 80 160
240
320 400 480
0
10
20
30
0
200
400
[°C]
[°F]
[bar]
[psi]
25020015010050
0
-50
≥ DN50
DN8…40
-80
0 80 160
240
320 400 480
0
10
20
30
40
0
[°C]
[°F]
[bar][psi]
25020015010050
0
-50
50
 33 With flange material 1.4301 (F304); wetted parts Alloy C22
Proline Promass F 500
A0028785-EN
Lap joint flange JIS B2220
 34 With flange material 1.4301 (F304); wetted parts Alloy C22
Thread DIN 11851
66 Endress+Hauser
 35 With connection material 1.4404 (316/316L)
DIN 11851 allows for applications up to +140 °C (+284 °F) if suitable sealing materials are used. Please take this into account when selecting seals and counterparts, as these components can limit the pressure and temperature range.
A0028786-EN
A0028794-EN
Page 67
Proline Promass F 500
DN8…40
≥ DN50
-80
0 80 160
240
320 400 480
0
10
20
30
40
0
[°C]
[°F]
[bar][psi]
25020015010050
0
-50
50
-80
0 80 160
240
320 400 480
0
10
20
0
[°C]
[°F]
[bar][psi]
25020015010050
0
-50
-80
0 80 160
240
320 400 480
0
10
20
0
[°C]
[°F]
[bar][psi]
25020015010050
0
-50
Thread DIN 11864-1 Form A
A0028798-EN
 36 With connection material 1.4404 (316/316L)
Thread ISO 2853
A0028799-EN
 37 With connection material 1.4404 (316/316L)
Thread SMS 1145
A0028800-EN
 38 With connection material 1.4404 (316/316L)
SMS 1145 allows for applications up to 16 bar (232 psi) if suitable sealing materials are used. Please take this into account when selecting seals and counterparts, as these components can limit the pressure and temperature range.
Endress+Hauser 67
Page 68
VCO
60
70
80
90
100
800
[bar][psi]
-80
0 80 160
240
320 400 480
[°C]
[°F]
25020015010050
0
-50
-80
0 80 160
240
320 400 480
0
10
20
0
[°C]
[°F]
[bar][psi]
25020015010050
0
-50
 39 With connection material 1.4404 (316/316L)
Tri-Clamp
Proline Promass F 500
A0028801-EN
The clamp connections are suitable up to a maximum pressure of 16 bar (232 psi). Please observe the operating limits of the clamp and seal used as they can be over 16 bar (232 psi). The clamp and seal are not included in the scope of supply.

Sensor housing

For standard versions with the temperature range –50 to +150 °C (–58 to +302 °F), the sensor housing is filled with dry nitrogen gas and protects the electronics and mechanics inside.
For all other temperature versions the sensor housing is filled with dry inert gas.
If a measuring tube fails (e.g. due to process characteristics like corrosive or abrasive fluids), the fluid will initially be contained by the sensor housing.
In the event of a tube failure, the pressure level inside the sensor housing will rise according to the operating process pressure. If the user judges that the sensor housing pressure rating/burst pressure does not provide an adequate safety margin, the device can be fitted with a rupture disk. This prevents excessively high pressure from forming inside the sensor housing. Therefore, the use of a rupture disk is strongly recommended in applications involving high gas pressures, and particularly in applications in which the process pressure is greater than 2/3 of the sensor housing burst pressure.
If there is a need to drain the leaking medium into a discharge device, the sensor should be fitted with a rupture disk. Connect the discharge to the additional threaded connection →  92.
If the sensor is to be purged with gas (gas detection), it should be equipped with purge connections.
Do not open the purge connections unless the containment can be filled immediately with a dry, inert gas. Use only low pressure to purge. Maximum pressure: 5 bar (72.5 psi).
68 Endress+Hauser
Sensor housing nominal pressure rating and burst pressure
The following sensor housing nominal pressure ratings/burst pressures are only valid for standard devices and/or devices equipped with closed purge connections (not opened/as delivered).
A0032216-EN
Page 69
Proline Promass F 500
If a device fitted with purge connections (order code for "Sensor option", option CH "Purge connection") is connected to the purge system, the maximum nominal pressure is determined by the purge system itself or by the device, depending on which component has the lower nominal pressure classification.
If the device is fitted with a rupture disk (order code for "Sensor option", option CA "Rupture disk"), the rupture disk trigger pressure is decisive for the maximum nominal pressure .
The sensor housing burst pressure refers to a typical internal pressure which is reached prior to mechanical failure of the sensor housing and which was determined during type testing. The corresponding type test declaration can be ordered with the device (order code for "Additional approval", option LN "Sensor housing burst pressure, type test").

Rupture disk

DN Sensor housing nominal
pressure
(designed with a safety factor
≥ 4)
[mm] [in] [bar] [psi] [bar] [psi]
8 ³⁄₈ 40 580 255 3 698
15 ½ 40 580 200 2 900
25 1 40 580 280 4 060
40 40 580 180 2 610
50 2 40 580 195 2 828
80 3 25 362 105 1 522
100 4 16 232 85 1 232
150 6 16 232 80 1 160
250 10 10 145 57 826
Sensor housing burst pressure
For information on the dimensions: see the "Mechanical construction" section→  73
To increase the level of safety, a device version with a rupture disk with a trigger pressure of 10 to 15 bar (145 to 217.5 psi)can be used (order code for "Sensor option", option CA "rupture disk").
The use of rupture disks cannot be combined with the separately available heating jacket.
Information on the dimensions of the rupture disk: →  92

Flow limit

Select the nominal diameter by optimizing between the required flow range and permissible pressure loss.
For an overview of the full scale values for the measuring range, see the "Measuring range" section →  12
• The minimum recommended full scale value is approx. 1/20 of the maximum full scale value
• In most applications, 20 to 50 % of the maximum full scale value can be considered ideal
• A low full scale value must be selected for abrasive media (such as liquids with entrained solids): flow velocity < 1 m/s (< 3 ft/s).
• For gas measurement the following rules apply: – The flow velocity in the measuring tubes should not exceed half the sound velocity (0.5 Mach). – The maximum mass flow depends on the density of the gas: formula →  12
To calculate the flow limit, use the Applicator sizing tool →  134

Pressure loss

To calculate the pressure loss, use the Applicator sizing tool →  134
Promass F with reduced pressure loss: order code for "Sensor option", option CE "Reduced pressure loss"

System pressure

It is important that cavitation does not occur, or that gases entrained in the liquids do not outgas. This is prevented by means of a sufficiently high system pressure.
Endress+Hauser 69
Page 70
For this reason, the following mounting locations are recommended:
• At the lowest point in a vertical pipe
• Downstream from pumps (no danger of vacuum)
Proline Promass F 500
A0028777

Thermal insulation

In the case of some fluids, it is important to keep the heat radiated from the sensor to the transmitter to a low level. A wide range of materials can be used for the required insulation.
The following device versions are recommended for versions with thermal insulation:
• Version with extended neck for insulation: Order code for "Sensor option", option CG with an extended neck length of 105 mm (4.13 in).
• Extended temperature version: Order code for "Measuring tube material", option SD, SE, SF or TH with an extended neck length of 105 mm (4.13 in).
• High-temperature version: Order code for "Measuring tube material", option TT or TU with an extended neck length of 142 mm (5.59 in).
NOTICE
Electronics overheating on account of thermal insulation!
Recommended orientation: horizontal orientation, sensor connection housing pointing
downwards. Do not insulate the sensor connection housing.
Maximum permissible temperature at the lower end of the sensor connection housing:
80 °C (176 °F) Thermal insulation with extended neck free: We recommend that you do not insulate the
extended neck in order to ensure optimum dissipation of heat.
A0034391
 40 Thermal insulation with extended neck free
Low-temperature version: It is generally not necessary to insulate the sensor connection housing. If insulation is provided, the rules that apply are the same as those for thermal insulation.

Heating

Some fluids require suitable measures to avoid loss of heat at the sensor.
Heating options
• Electrical heating, e.g. with electric band heaters
• Via pipes carrying hot water or steam
• Via heating jackets
Heating jackets for the sensors can be ordered as accessories from Endress+Hauser .→  133
70 Endress+Hauser
Page 71
Proline Promass F 500
NOTICE
Danger of overheating when heating
Ensure that the temperature at the lower end of the transmitter housing does not exceed
80 °C (176 °F). Ensure that sufficient convection takes place at the transmitter neck.
Ensure that a sufficiently large area of the transmitted neck remains exposed. The uncovered
part serves as a radiator and protects the electronics from overheating and excessive cooling. If using in potentially explosive atmospheres, observe the information in the device-specific Ex
documentation. For detailed information on the temperature tables, see the separate document entitled "Safety Instructions" (XA) for the device.

Vibrations

The high oscillation frequency of the measuring tubes ensures that the correct operation of the measuring system is not influenced by plant vibrations.
Endress+Hauser 71
Page 72
Proline Promass F 500

Custody transfer measurement

The measuring device is optionally tested in accordance with OIML R117/R81 and has an EU type evaluation certificate which authorizes the use in EU type-examination certificates according to Measuring Instruments Directive 2014/32/EU for service subject to legal metrological control ("custody transfer") for liquids other than water and cryogenic liquids (Annex VII).
The permitted fluid temperature in these applications is –200 to +90 °C (–328 to +194 °F).
The measuring device is optionally tested in accordance with OIML R137 and has an EU type­examination certificate according to Measuring Instruments Directive 2014/32/EU for service as a gas meter subject to legal metrological control ("custody transfer") (Annex IV). The permitted fluid temperature in these applications is –25 to +55 °C (–13 to +131 °F).
The device is used with a legally controlled totalizer on the local display and optionally with legally controlled outputs.
Measuring devices subject to legal metrological control totalize in both directions, i.e. all the outputs consider flow components in the positive (forward) and negative (reverse) flow direction.
Generally a measuring device subject to legal metrological control is secured against tampering by seals on the transmitter or sensor. These seals may normally only be opened by a representative of the competent authority for legal metrology controls.
After putting the device into circulation or after sealing the device, operation is only possible to a limited extent.
Detailed ordering information is available from your local Endress+Hauser sales center for national approvals, which are based on the OIML certificates, for applications with liquids other than water, cryogenic liquids or gases.
72 Endress+Hauser
Page 73
Proline Promass F 500
N
P
F
G
A
Q
T
G
A
B
C
Q
Q
F

Mechanical construction

Dimensions in SI units Housing of Proline 500 – digital transmitter

Non-hazardous area or hazardous area: Zone 2; Class I, Division 2
A0033789
Order code for "Transmitter housing", option A "Aluminum, coated" and order code for "Integrated ISEM electronics", option A "Sensor"
A
[mm]
167 232 80 187 24 21
F
[mm]
G
[mm]
N
[mm]
P
[mm]
Q
[mm]
Order code for "Transmitter housing", option D "Polycarbonate" and order code for "Integrated ISEM electronics", option A "Sensor"
A
[mm]
177 234 90 197 17 22
F
[mm]
G
[mm]
N
[mm]
P
[mm]
Q
[mm]
Housing of Proline 500 transmitter
Hazardous area: Zone 2; Class I, Division 2 or Zone 1; Class I, Division 1
Order code for "Transmitter housing", option A "Aluminum, coated" and order code for "Integrated ISEM
A0033788
electronics", option B "Transmitter"
A
[mm]
188 85 103 318 217 130 239
B
[mm]
C
[mm]
F
[mm]
G
[mm]
Q
[mm]
T
[mm]
Endress+Hauser 73
Page 74
Proline Promass F 500
A
B
C
G
D
L
E
F
K
M
Order code for "Transmitter housing", option L "Cast, stainless" and order code for "Integrated ISEM electronics", option B "Transmitter"
A
[mm]
B
[mm]
C
[mm]
F
[mm]
G
[mm]
Q
[mm]
[mm]
188 85 103 295 217 130 239
Sensor with connection housing
Order code for "Sensor connection housing", option A "Aluminum, coated"
DN A
1)
[mm] [mm] [mm] [mm] [mm] [mm] [mm] [mm] [mm] [mm] [mm]
8 148 94 54 75 185 260 136 5.35
15 148 94 54 75 185 260 136 8.30
25 148 94 54 75 185 260 136 12.0
40 148 94 54 105 189.5 294.5 136 17.6
50 148 94 54 141 199.5 340.5 136 26.0
80 148 94 54 200 219.5 419.5 136 40.5
100 148 94 54 254 238 492 136 51.2
150 148 94 54 378 259 637 136 68.9
250 148 94 54 548 302.5 850.5 136 102.3
1)
B
C D E
2) 3)
F
2) 3)
G K L M
4)
4)
4)
4)
4)
4)
4)
4)
4)
T
A0033787
70
70
70
79
99
139
176
218
305
1) Depending on the cable gland used: values up to + 30 mm
2) With order code for "Sensor option", option CG or order code for "Measuring tube material", option SD, SE,
SF, TH, LA: values +70 mm
3) With order code for "Measuring tube material", option TT, TU: values +104 mm
4) Depends on the process connection in question →  76
Order code for "Sensor connection housing", option B "Stainless"
DN A
[mm] [mm] [mm] [mm] [mm] [mm] [mm] [mm] [mm] [mm] [mm]
1)
8 137 78 59 75 180 255 134 5.35
15 137 78 59 75 180 255 134 8.30
25 137 78 59 75 180 255 134 12.0
40 137 78 59 105 184.5 289.5 134 17.6
50 137 78 59 141 194.5 335.5 134 26.0
1)
B
C D E
2) 3)
F
2) 3)
G K L M
4)
4)
4)
4)
4)
70
70
70
79
99
74 Endress+Hauser
Page 75
Proline Promass F 500
DN A
1)
1)
B
C D E
2) 3)
F
2) 3)
G K L M
[mm] [mm] [mm] [mm] [mm] [mm] [mm] [mm] [mm] [mm] [mm]
80 137 78 59 200 214.5 414.5 134 40.5
100 137 78 59 254 233 487 134 51.2
150 137 78 59 378 254 632 134 68.9
250 137 78 59 548 297.5 845.5 134 102.3
4)
4)
4)
4)
139
176
218
305
1) Depending on the cable gland used: values up to + 30 mm
2) With order code for "Sensor option", option CG or order code for "Measuring tube material", option SD, SE, SF, TH, LA: values +70 mm
3) With order code for "Measuring tube material", option TT, TU: values +104 mm
4) Depends on the process connection in question →  76
Order code for "Sensor connection housing", option C "Ultra-compact hygienic, stainless"
DN A
1)
[mm] [mm] [mm] [mm] [mm] [mm] [mm] [mm] [mm] [mm]
8 124 68 56 75 180 255 112 5.35
15 124 68 56 75 180 255 112 8.30
25 124 68 56 75 180 255 112 12.0
40 124 68 56 105 184.5 289.5 112 17.6
50 124 68 56 141 194.5 335.5 112 26.0
80 124 68 56 200 214.5 414.5 112 40.5
100 124 68 56 254 233 487 112 51.2
150 124 68 56 378 254 632 112 68.9
250 124 68 56 548 297.5 845.5 112 102.3
1)
B
C D E
2) 3)
F
2) 3)
G K L M
4)
4)
4)
4)
4)
4)
4)
4)
4)
70
70
70
79
99
139
176
218
305
1) Depending on the cable gland used: values up to + 30 mm
2) With order code for "Sensor option", option CG or order code for "Measuring tube material", option SD, SE, SF, TH, LA: values +70 mm
3) With order code for "Measuring tube material", option TT, TU: values +104 mm
4) Depends on the process connection in question →  76
Order code for "Sensor connection housing", option L "Cast, stainless"
DN A
1)
[mm] [mm] [mm] [mm] [mm] [mm] [mm] [mm] [mm] [mm] [mm]
8 145 86 59 75 208 283 136 5.35
15 145 86 59 75 208 283 136 8.30
25 145 86 59 75 208 283 136 12.0
40 145 86 59 105 212.5 317.5 136 17.6
50 145 86 59 141 222.5 363.5 136 26.0
80 145 86 59 200 242.5 442.5 136 40.5
100 145 86 59 254 261 515 136 51.2
150 145 86 59 378 282 660 136 68.9
250 145 86 59 548 325.5 873.5 136 102.3
1) Depending on the cable gland used: values up to + 30 mm
2) With order code for "Sensor option", option CG or order code for "Measuring tube material", option SD, SE, SF, TH, LA: values +70 mm
3) With order code for "Measuring tube material", option TT, TU: values +104 mm
4) Depends on the process connection in question →  76
1)
B
C D E
2) 3)
F
2) 3)
G K L M
4)
4)
4)
4)
4)
4)
4)
4)
4)
70
70
70
79
99
139
176
218
305
Endress+Hauser 75
Page 76
Flange connections
C
D
A
B
E
L
Fixed flange EN 1092-1, ASME B16.5, JIS B2220
Length tolerance for dimension L in mm:
• DN ≤ 100: +1.5 / –2.0
• DN ≥ 125: +3.5
Proline Promass F 500
A0015621
Flange according to EN 1092-1 (DIN 2501): PN16
1.4404 (F316/F316L): order code for "Process connection", option D1S Alloy C22: order code for "Process connection", option D1C
Flange with groove according to EN 1092-1 Form D (DIN 2512N): PN16
1.4404 (F316/F316L): order code for "Process connection", option D5S Alloy C22: order code for "Process connection", option D5C
DN
[mm]
100 220 180 8 × ⌀18 20 107.1 1 127/1 400
150 285 240 8 × ⌀22 22 159.3 1 330/1 700
250 405 355 12 × ⌀26 26 260.4 1 775
Surface roughness (flange): EN 1092-1 Form B1 (DIN 2526 Form C), Ra 3.2 to 12.5 µm
1) Installation length in accordance with NAMUR recommendation NE 132 optionally available (order code for "Process connection", option D1N or D5N (with groove))
Flange according to EN 1092-1 (DIN 2501): PN16 with reduction in nominal diameter
1.4404 (F316/F316L
DN
[mm]
100 80 DHS 220 180 8 × ⌀ 18 20 107.1 874
150 100 DJS 285 240 8 × ⌀ 22 22 159.3 1 167
200 150 DLS 340 295 12 × ⌀ 22 24 206.5 1 461
Surface roughness (flange): EN 1092-1 Form B1 (DIN 2526 Form C), Ra 3.2 to 12.5 µm
A
[mm]
reduction
to DN [mm]
B
[mm]
Order code for
"Process
connection",
option
C
[mm]
A
[mm]B[mm]
D
[mm]
C
[mm]
E
[mm]
D
[mm]E[mm]L[mm]
L
[mm]
1)
1)
76 Endress+Hauser
Page 77
Proline Promass F 500
Flange according to EN 1092-1 (DIN 2501): PN 40
1.4404 (F316/F316L): order code for "Process connection", option D2S Alloy C22: order code for "Process connection", option D2C
Flange with groove according to EN 1092-1 Form D (DIN 2512N): PN 40
1.4404 (F316/F316L): order code for "Process connection", option D6S Alloy C22: order code for "Process connection", option D6C
DN
[mm]
1)
8
A
[mm]
B
[mm]
C
[mm]
D
[mm]
E
[mm]
95 65 4 × ⌀14 16 17.3 370/510
15 95 65 4 × ⌀14 16 17.3 404/510
25 115 85 4 × ⌀14 18 28.5 440/600
40 150 110 4 × ⌀18 18 43.1 550
50 165 125 4 × ⌀18 20 54.5 715/715
80 200 160 8 × ⌀18 24 82.5 840/915
100 235 190 8 × ⌀22 24 107.1 1 127
150 300 250 8 × ⌀26 28 159.3 1 370
250 450 385 12 × ⌀33 38 258.8 1 845
Surface roughness (flange): EN 1092-1 Form B1 (DIN 2526 Form C), Ra 3.2 to 12.5 µm
L
[mm]
2)
2)
2)
2)
2)
1) DN 8 with DN 15 flanges as standard
2) Installation length in accordance with NAMUR recommendation NE 132 optionally available (order code for "Process connection", option D2N or D6N (with groove))
Flange according to EN 1092-1 (DIN 2501): PN 40 (with DN 25 flanges)
1.4404 (F316/F316L): order code for "Process connection", option R2S
DN
[mm]
A
[mm]
B
[mm]
C
[mm]
D
[mm]
E
[mm]
L
[mm]
8 115 85 4 × ⌀14 18 28.5 440
15 115 85 4 × ⌀14 18 28.5 440
Surface roughness (flange): EN 1092-1 Form B1 (DIN 2526 Form C), Ra 3.2 to 12.5 µm
Flange according to EN 1092-1 (DIN 2501): PN 40 with reduction in nominal diameter
1.4404 (F316/F316L)
DN
[mm]
reduction
to DN [mm]
Order code for
"Process
connection",
A
[mm]B[mm]
C
[mm]
D
[mm]E[mm]L[mm]
option
50 40 DFS 165 125 4 × ⌀ 18 20 54.5 555
80 50 DGS 200 160 8 × ⌀ 18 24 82.5 840
100 80 DIS 235 190 8 × ⌀ 22 24 107.1 874
150 100 DKS 300 250 8 × ⌀ 26 28 159.3 1 167
200 150 DMS 375 320 12 × ⌀ 30 34 206.5 1 461
Surface roughness (flange): EN 1092-1 Form B1 (DIN 2526 Form C), Ra 3.2 to 12.5 µm
Endress+Hauser 77
Page 78
Proline Promass F 500
Flange according to EN 1092-1 (DIN 2501): PN 63
1.4404 (F316/F316L): order code for "Process connection", option D3S Alloy C22: order code for "Process connection", option D3C
Flange with groove according to EN 1092-1 Form D (DIN 2512N): PN 63
1.4404 (F316/F316L): order code for "Process connection", option D7S Alloy C22: order code for "Process connection", option D7C
DN
[mm]
50 180 135 4 × ⌀22 26 54.5 724
80 215 170 8 × ⌀22 28 81.7 875
100 250 200 8 × ⌀26 30 106.3 1 127
150 345 280 8 × ⌀33 36 157.1 1 410
250 470 400 12 × ⌀36 46 255.4 1 885
Surface roughness (flange): EN 1092-1 Form B1 (DIN 2526 Form C), Ra 3.2 to 12.5 µm EN 1092-1 Form B2 (DIN 2526 Form E), Ra
0.8 to 3.2 µm
Flange according to EN 1092-1 (DIN 2501): PN 100
1.4404 (F316/F316L): order code for "Process connection", option D4S Alloy C22: order code for "Process connection", option D4C
Flange with groove according to EN 1092-1 Form D (DIN 2512N): PN 100
1.4404 (F316/F316L): order code for "Process connection", option D8S Alloy C22: order code for "Process connection", option D8C
DN
[mm]
1)
8
15 105 75 4 × ⌀14 20 17.3 420
25 140 100 4 × ⌀18 24 28.5 470
40 170 125 4 × ⌀22 26 42.5 590
50 195 145 4 × ⌀26 28 53.9 740
80 230 180 8 × ⌀26 32 80.9 885
100 265 210 8 × ⌀30 36 104.3 1 127
150 355 290 12 × ⌀33 44 154.0 1 450
Surface roughness (flange): EN 1092-1 Form B2 (DIN 2526 Form E), Ra 0.8 to 3.2 µm
A
[mm]
A
[mm]
105 75 4 × ⌀14 20 17.3 400
B
[mm]
B
[mm]
C
[mm]
C
[mm]
D
[mm]
D
[mm]
E
[mm]
E
[mm]
[mm]
[mm]
L
L
1) DN 8 with DN 15 flanges as standard
Flange according to EN 1092-1 (DIN 2501): PN 100 Alloy C22: order code for "Process connection", option D4C
Flange with groove according to EN 1092-1 Form D (DIN 2512N): PN 100 Alloy C22: order code for "Process connection", option D8C
DN
[mm]
250 505 430 12 × ⌀39 60 248.0 1 949
Surface roughness (flange): EN 1092-1 Form B2 (DIN 2526 Form E), Ra 0.8 to 3.2 µm
A
[mm]
B
[mm]
C
[mm]
D
[mm]
E
[mm]
L
[mm]
78 Endress+Hauser
Page 79
Proline Promass F 500
Flange according to ASME B16.5: Class 150
1.4404 (F316/F316L): order code for "Process connection", option AAS Alloy C22: order code for "Process connection", option AAC
DN
[mm]
1)
8
A
[mm]
B
[mm]
C
[mm]
D
[mm]
E
[mm]
90 60.3 4 × ⌀15.7 11.2 15.7 370
15 90 60.3 4 × ⌀15.7 11.2 15.7 404
25 110 79.4 4 × ⌀15.7 14.2 26.7 440
40 125 98.4 4 × ⌀15.7 17.5 40.9 550
50 150 120.7 4 × ⌀19.1 19.1 52.6 715
80 190 152.4 4 × ⌀19.1 23.9 78.0 840
100 230 190.5 8 × ⌀19.1 23.9 102.4 1 127
150 280 241.3 8 × ⌀22.4 25.4 154.2 1 398
250 405 362 12 × ⌀25.4 30.2 254.5 1 832
Surface roughness (flange): Ra 3.2 to 6.3 µm
1) DN 8 with DN 15 flanges as standard
L
[mm]
Flange according to ASME B16.5: Class 150 with reduction in nominal diameter
1.4404 (F316/F316L)
DN
[mm]
reduction
to DN [mm]
Order code for
"Process
connection",
A
[mm]B[mm]
C
[mm]
D
[mm]E[mm]L[mm]
option
50 40 AHS 150 120.7 4 × ⌀ 19.1 19.1 52.6 550
80 50 AJS 190 152.4 4 × ⌀ 19.1 23.9 78.0 720
100 80 ALS 230 190.5 8 × ⌀ 19.1 23.9 102.4 874
150 100 ANS 280 241.3 8 × ⌀ 22.4 25.4 154.2 1 167
200 150 APS 345 298.5 8 × ⌀ 22.4 29 202.7 1 461
Surface roughness (flange): Ra 3.2 to 6.3 µm
Flange according to ASME B16.5: Class 300
1.4404 (F316/F316L): order code for "Process connection", option ABS Alloy C22: order code for "Process connection", option ABC
DN
[mm]
1)
8
A
[mm]
B
[mm]
C
[mm]
D
[mm]
E
[mm]
95 66.7 4 × ⌀15.7 14.2 15.7 370
[mm]
15 95 66.7 4 × ⌀15.7 14.2 15.7 404
25 125 88.9 4 × ⌀19.1 17.5 26.7 440
40 155 114.3 4 × ⌀22.3 20.6 40.9 550
50 165 127 8 × ⌀19.1 22.3 52.6 715
80 210 168.3 8 × ⌀22.3 28.4 78.0 840
100 255 200 8 × ⌀22.3 31.7 102.4 1 127
150 320 269.9 12 × ⌀22.3 36.5 154.2 1 417
250 445 387.4 16 × ⌀28.4 47.4 254.5 1 863
Surface roughness (flange): Ra 3.2 to 6.3 µm
L
1) DN 8 with DN 15 flanges as standard
Endress+Hauser 79
Page 80
Proline Promass F 500
Flange according to ASME B16.5: Class 300 with reduction in nominal diameter
1.4404 (F316/F316L)
DN
[mm]
50 40 AIS 165 127 8 × ⌀ 19.1 22.3 52.6 615
80 50 AKS 210 168.3 8 × ⌀ 22.3 28.4 78.0 732
100 80 AMS 255 200 8 × ⌀ 22.3 31.7 102.4 894
150 100 AOS 320 269.9 12 × ⌀ 22.3 36.5 154.2 1 187
200 150 AQS 380 330.2 12 × ⌀ 25.4 41.7 202.7 1 461
Surface roughness (flange): Ra 3.2 to 6.3 µm
Flange according to ASME B16.5: Class 600
1.4404 (F316/F316L): order code for "Process connection", option ACS Alloy C22: order code for "Process connection", option ACC
Surface roughness (flange): Ra 3.2 to 6.3 µm
reduction
to DN [mm]
DN
[mm]
1)
8
15 95 66.7 4 × ⌀15.7 20.6 13.9 420
25 125 88.9 4 × ⌀19.1 23.9 24.3 490
40 155 114.3 4 × ⌀22.3 28.7 38.1 600
50 165 127 8 × ⌀19.1 31.8 49.2 742
80 210 168.3 8 × ⌀22.3 38.2 73.7 900
100 275 215.9 8 × ⌀25.4 48.4 97.3 1 157
150 355 292.1 12 × ⌀28.4 47.8 154.2 1 467
250 510 431.8 16 × ⌀35.1 69.9 254.5 1 946
[mm]
Order code for
"Process
connection",
option
A
95 66.7 4 × ⌀15.7 20.6 13.9 400
B
[mm]
A
[mm]B[mm]
C
[mm]
C
[mm]
D
[mm]
D
[mm]E[mm]L[mm]
E
[mm]
[mm]
L
1) DN 8 with DN 15 flanges as standard
Flange JIS B2220: 10K
1.4404 (F316/F316L): order code for "Process connection", option NDS Alloy C22: order code for "Process connection", option NDC
DN
[mm]
50 155 120 4 × ⌀19 16 50 715
80 185 150 8 × ⌀19 18 80 832
100 210 175 8 × ⌀19 18 100 1 127
150 280 240 8 × ⌀23 22 150 1 354
250 400 355 12 × ⌀25 24 250 1 775
Surface roughness (flange): Ra 3.2 to 6.3 µm
A
[mm]
B
[mm]
C
[mm]
D
[mm]
E
[mm]
L
[mm]
80 Endress+Hauser
Page 81
Proline Promass F 500
Flange JIS B2220: 20K
1.4404 (F316/F316L): order code for "Process connection", option NES Alloy C22: order code for "Process connection", option NEC
DN
[mm]
1)
8
A
[mm]
B
[mm]
C
[mm]
D
[mm]
E
[mm]
95 70 4 × ⌀15 14 15 370
15 95 70 4 × ⌀15 14 15 404
25 125 90 4 × ⌀19 16 25 440
40 140 105 4 × ⌀19 18 40 550
50 155 120 8 × ⌀19 18 50 715
80 200 160 8 × ⌀23 22 80 832
100 225 185 8 × ⌀23 24 100 1 127
150 305 260 12 × ⌀25 28 150 1 386
250 430 380 12 × ⌀27 34 250 1 845
Surface roughness (flange): Ra 1.6 to 3.2 µm
1) DN 8 with DN 15 flanges as standard
L
[mm]
Flange JIS B2220: 40K
1.4404 (F316/F316L): order code for "Process connection", option NGS Alloy C22: order code for "Process connection", option NGC
DN
[mm]
1)
8
A
[mm]
B
[mm]
C
[mm]
D
[mm]
E
[mm]
115 80 4 × ⌀19 20 15 400
15 115 80 4 × ⌀19 20 15 425
25 130 95 4 × ⌀19 22 25 485
40 160 120 4 × ⌀23 24 38 600
50 165 130 8 × ⌀19 26 50 760
80 210 170 8 × ⌀23 32 75 890
100 250 205 8 × ⌀25 36 100 1 167
150 355 295 12 × ⌀33 44 150 1 498
Surface roughness (flange): Ra 1.6 to 3.2 µm
1) DN 8 with DN 15 flanges as standard
Flange JIS B2220: 63K
1.4404 (F316/F316L): order code for "Process connection", option NHS Alloy C22: order code for "Process connection", option NHC
DN
[mm]
1)
8
A
[mm]
B
[mm]
C
[mm]
D
[mm]
E
[mm]
120 85 4 × ⌀19 23 12 420
15 120 85 4 × ⌀19 23 12 440
25 140 100 4 × ⌀23 27 22 494
40 175 130 4 × ⌀25 32 35 620
50 185 145 8 × ⌀23 34 48 775
80 230 185 8 × ⌀25 40 73 915
100 270 220 8 × ⌀27 44 98 1 167
L
[mm]
L
[mm]
Endress+Hauser 81
Page 82
Proline Promass F 500
Flange JIS B2220: 63K
1.4404 (F316/F316L): order code for "Process connection", option NHS Alloy C22: order code for "Process connection", option NHC
DN
[mm]
150 365 305 12 × ⌀33 54 146 1 528
Surface roughness (flange): Ra 1.6 to 3.2 µm
1) DN 8 with DN 15 flanges as standard
A
[mm]
B
[mm]
C
[mm]
D
[mm]
E
[mm]
L
[mm]
82 Endress+Hauser
Page 83
Proline Promass F 500
C
D
A
B
E
X X
L
Fixed flange DIN 11864-2
A0015627
 41 Detail X: Asymmetrical process connection; the part shown in blue is provided by the supplier.
Length tolerance for dimension L in mm: +1.5 / –2.0
Flange DIN11864-2 Form A, for pipe according to DIN11866 series A, flat with notch
1.4404 (316/316L) Order code for "Process connection", option KCS
DN
[mm]
8 54 37 4 × ⌀9 10 10 387
15 59 42 4 × ⌀9 10 16 418
25 70 53 4 × ⌀9 10 26 454
40 82 65 4 × ⌀9 10 38 560
50 94 77 4 × ⌀9 10 50 720
80 133 112 8 × ⌀11 12 81 900
100 159 137 8 × ⌀11 14 100 1 127
3A-version available: order code for "Additional approval", option LP in conjunction with Ra ≤ 0.8 µm: order code for "Measuring tube material", option SB, SE or Ra ≤ 0.4 µm: order code for "Measuring tube material", option SC, SF
A
[mm]
B
[mm]
C
[mm]
D
[mm]
E
[mm]
L
[mm]
Endress+Hauser 83
Page 84
Proline Promass F 500
L
C
D
A
B
F
E
Lap joint flange EN 1092-1, ASME B16.5, JIS B2220
Length tolerance for dimension L in mm: +1.5 / –2.0
Lap joint flange according to EN 1092-1 Form D: PN 40
1.4301 (F304), wetted parts Alloy C22 Order code for "Process connection", option DAC
DN
[mm]A[mm]B[mm]
2)
8
95 65 4 × ⌀ 14 14.5 45 17.3 370 0
15 95 65 4 × ⌀ 14 14.5 45 17.3 404 0
25 115 85 4 × ⌀ 14 16.5 68 28.5 444 +4
40 150 110 4 × ⌀ 18 21 88 43.1 560 +10
50 165 125 4 × ⌀ 18 23 102 54.5 719 +4
80 200 160 8 × ⌀ 18 29 138 82.5 848 +8
100 235 190 8 × ⌀ 22 34 162 107.1 1 131 +4
Surface roughness (flange): Ra 3.2 to 12.5 µm
C
[mm]
D
[mm]E[mm]F[mm]L[mm]
A0022221
L
diff
[mm]
1)
1) Difference to installation length of the welding neck flange (order code for "Process connection", option D2C)
2) DN 8 with DN 15 flanges as standard
Lap joint flange according to ASME B16.5: Class 150
1.4301 (F304), wetted parts Alloy C22 Order code for "Process connection", option ADC
DN
[mm]A[mm]B[mm]
2)
8
90 60.3 4 × ⌀ 15.7 15 35.1 15.7 370 0
C
[mm]
D
[mm]E[mm]F[mm]L[mm]
L
diff
[mm]
1)
15 90 60.3 4 × ⌀ 15.7 15 35.1 15.7 404 0
25 110 79.4 4 × ⌀ 15.7 16 50.8 26.7 440 0
40 125 98.4 4 × ⌀ 15.7 15.9 73.2 40.9 550 0
50 150 120.7 4 × ⌀ 19.1 19 91.9 52.6 715 0
80 190 152.4 4 × ⌀ 19.1 22.3 127.0 78.0 840 0
84 Endress+Hauser
Page 85
Proline Promass F 500
Lap joint flange according to ASME B16.5: Class 150
1.4301 (F304), wetted parts Alloy C22 Order code for "Process connection", option ADC
DN
[mm]A[mm]B[mm]
C
[mm]
D
[mm]E[mm]F[mm]L[mm]
L
diff
[mm]
100 230 190.5 8 × ⌀ 19.1 26 157.2 102.4 1 127 0
Surface roughness (flange): Ra 3.2 to 12.5 µm
1) Difference to installation length of the welding neck flange (order code for "Process connection", option AAC)
2) DN 8 with DN 15 flanges as standard
Lap joint flange according to ASME B16.5: Class 300
1.4301 (F304), wetted parts Alloy C22 Order code for "Process connection", option AEC
DN
[mm]A[mm]B[mm]
2)
8
95 66.7 4 × ⌀ 15.7 16.5 35.1 15.7 376 +6
C
[mm]
D
[mm]E[mm]F[mm]L[mm]
L
diff
[mm]
15 95 66.7 4 × ⌀ 15.7 16.5 35.1 15.7 406 +2
25 125 88.9 4 × ⌀ 19.1 21.0 50.8 26.7 450 +10
40 155 114.3 4 × ⌀ 22.3 23.0 73.2 40.9 564 +14
50 165 127 8 × ⌀ 19.1 25.5 91.9 52.6 717 +2
80 210 168.3 8 × ⌀ 22.3 31.0 127.0 78.0 852.6 +12.6
100 255 200 8 × ⌀ 22.3 32.0 157.2 102.4 1 139 +12
Surface roughness (flange): Ra 3.2 to 12.5 µm
1)
1)
1) Difference to installation length of the welding neck flange (order code for "Process connection", option ABC)
2) DN 8 with DN 15 flanges as standard
Lap joint flange according to ASME B16.5: Class 600
1.4301 (F304), wetted parts Alloy C22 Order code for "Process connection", option AFC
DN
[mm]A[mm]B[mm]
2)
8
95 66.7 4 × ⌀ 15.7 17.0 35.1 13.9 400 0
C
[mm]
D
[mm]E[mm]F[mm]L[mm]
L
diff
[mm]
15 95 66.7 4 × ⌀ 15.7 17.0 35.1 13.9 420 0
25 125 88.9 4 × ⌀ 19.1 21.5 50.8 24.3 490 0
40 155 114.3 4 × ⌀ 22.3 25.0 73.2 38.1 600 0
50 165 127 8 × ⌀ 19.1 28.0 91.9 49.2 742 0
80 210 168.3 8 × ⌀ 22.3 35.0 127.0 73.7 900 0
100 275 215.9 8 × ⌀ 25.4 44.0 157.2 97.3 1 167 +10
Surface roughness (flange): Ra 3.2 to 12.5 µm
1) Difference to installation length of the welding neck flange (order code for "Process connection", option ACC)
2) DN 8 with DN 15 flanges as standard
1)
Endress+Hauser 85
Page 86
Proline Promass F 500
Lap joint flange JIS B2220: 20K
1.4301 (F304), wetted parts Alloy C22 Order code for "Process connection", option NIC
DN
[mm]
2)
8
15 95 70 4 × ⌀ 15 14 51 15 404 0
25 125 90 4 × ⌀ 19 18.5 67 25 440 0
40 140 105 4 × ⌀ 19 18.5 81 40 550 0
50 155 120 8 × ⌀ 19 23 96 50 715 0
80 200 160 8 × ⌀ 23 29 132 80 844 +12
100 225 185 8 × ⌀ 23 29 160 100 1 127 0
Surface roughness (flange): Ra 3.2 to 12.5 µm
1) Difference to installation length of the welding neck flange (order code for "Process connection", option NEC)
2) DN 8 with DN 15 flanges as standard
A
[mm]
B
[mm]
C
[mm]
D
[mm]
E
[mm]
F
[mm]
L
[mm]
95 70 4 × ⌀ 15 14 51 15 370 0
L
diff
[mm]
1)
86 Endress+Hauser
Page 87
Proline Promass F 500
L
A
B
Clamp connections
Tri-Clamp
Length tolerance for dimension L in mm: +1.5 / –2.0
Tri-Clamp (½"), for pipe according to DIN 11866 series C
1.4404 (316/316L) Order code for "Process connection", option FDW
DN
[mm]
8 ½ 25.0 9.5 367
15 ½ 25.0 9.5 398
3-A version available: order code for "Additional approval", option LP in conjunction with Ra ≤ 0.8 µm: order code for "Measuring tube material", option SB, SE or Ra ≤ 0.4 µm: order code for "Measuring tube material", option SC, SF
Clamp
[in]
A
[mm]
B
[mm]
A0015625
L
[mm]
Tri-Clamp ( ≥ 1"), for pipe according to DIN 11866 series C
1.4404 (316/316L) Order code for "Process connection", option FTS
DN
[mm]
8 1 50.4 22.1 367
15 1 50.4 22.1 398
25 1 50.4 22.1 434
40 50.4 34.8 560
50 2 63.9 47.5 720
80 3 90.9 72.9 900
100 4 118.9 97.4 1 127
3-A version available: order code for "Additional approval", option LP in conjunction with Ra ≤ 0.8 µm: order code for "Measuring tube material", option SB, SE or Ra ≤ 0.4 µm: order code for "Measuring tube material", option SC, SF
Clamp
[in]
A
[mm]
B
[mm]
L
[mm]
Endress+Hauser 87
Page 88
Threaded couplings
L
A
B
Thread DIN 11851, DIN11864-1, SMS 1145
Length tolerance for dimension L in mm: +1.5 / –2.0
Thread DIN 11851, for pipe according to DIN11866, series A
1.4404 (316/316L) Order code for "Process connection", option FMW
DN
[mm]
8 Rd 34 × ¹⁄₈ 16 367
15 Rd 34 × ¹⁄₈ 16 398
25 Rd 52 × ¹⁄₆ 26 434
40 Rd 65 × ¹⁄₆ 38 560
50 Rd 78 × ¹⁄₆ 50 720
80 Rd 110 × ¹⁄₄ 81 900
100 Rd 130 × ¹⁄₄ 100 1 127
3-A version available: order code for "Additional approval", option LP in conjunction with Ra ≤ 0.8 µm: order code for "Measuring tube material", option SB, SE
A
[in]
B
[mm]
Proline Promass F 500
A0015628
L
[mm]
Thread DIN11864-1 Form A, for pipe according to DIN11866, series A
1.4404 (316/316L) Order code for "Process connection", option FLW
DN
[mm]
8 Rd 28 × ¹⁄₈ 10 367
15 Rd 34 × ¹⁄₈ 16 398
25 Rd 52 × ¹⁄₈ 26 434
40 Rd 65 × ¹⁄₆ 38 560
50 Rd 78 × ¹⁄₆ 50 720
80 Rd 110 × ¹⁄₄ 81 900
100 Rd 130 × ¹⁄₄ 100 1 127
3-A version available: order code for "Additional approval", option LP in conjunction with Ra ≤ 0.8 µm: order code for "Measuring tube material", option SB, SE or Ra ≤ 0.4 µm: order code for "Measuring tube material", option SC, SF
A
[in]
B
[mm]
L
[mm]
88 Endress+Hauser
Page 89
Proline Promass F 500
Thread SMS 1145
1.4404 (316/316L) Order code for "Process connection", option SCS
DN
[mm]
8 Rd 40 × ¹⁄₆ 22.6 367
15 Rd 40 × ¹⁄₆ 22.6 398
25 Rd 40 × ¹⁄₆ 22.6 434
40 Rd 60 × ¹⁄₆ 35.6 560
50 Rd 70 × ¹⁄₆ 48.6 720
80 Rd 98 × ¹⁄₆ 72.9 900
100 Rd 132 × ¹⁄₆ 97.6 1 127
3-A version available: order code for "Additional approval", option LP in conjunction with Ra ≤ 0.8 µm: order code for "Measuring tube material", option SB, SE
A
[in]
B
[mm]
L
[mm]
Endress+Hauser 89
Page 90
Thread ISO 2853
L
A
B
Length tolerance for dimension L in mm: +1.5 / –2.0
Thread ISO 2853, for pipe according to ISO 2037
1.4404 (316/316L) Order code for "Process connection", option JSF
DN
[mm]
8 37.13 22.6 367
15 37.13 22.6 398
25 37.13 22.6 434
40 52.68 35.6 560
50 64.16 48.6 720
80 91.19 72.9 900
100 118.21 97.6 1 127
3-A version available: order code for "Additional approval", option LP in conjunction with Ra ≤ 0.8 µm: order code for "Measuring tube material", option SB, SE or Ra ≤ 0.4 µm: order code for "Measuring tube material", option SC, SF
1)
A
[mm]
B
[mm]
Proline Promass F 500
A0015623
L
[mm]
1) Max. thread diameter as per ISO 2853 annex A
90 Endress+Hauser
Page 91
Proline Promass F 500
L
A
B
VCO
Length tolerance for dimension L in mm: +1.5 / –2.0
8-VCO-4 (½")
1.4404 (316/316L) Order code for "Process connection", option CVS
DN
[mm]
8 AF 1 10.2 390
A
[in]
B
[mm]
A0015624
L
[mm]
12-VCO-4 (¾")
1.4404 (316/316L) Order code for "Process connection", option CWS
DN
[mm]
15 AF 1½ 15.7 430
A
[in]
B
[mm]
L
[mm]
Endress+Hauser 91
Page 92
Accessories
DN 8 ( ")...150 (6")!⁄# DN 250 (10")
L L
45°
A
A
SW 1"
2
½" NPT
1
½" NPT
35 (1.38) ca. 75 (approx. 2.95)
1
1;2
1
1;2
Rupture disk/purge connections
Proline Promass F 500
 42
1 Connection nipple for purge connections: order code for "Sensor options", option CH "Purge connection" 2 Connection nipple with rupture disk: order code for "Sensor option", option CA "Rupture disk"
DN A L
[mm] [mm] [mm]
8 62 216
15 62 220
25 62 260
40 67 310
50 79 452
80 101 560
100 120 684
150 141 880
250 182 380
A0028914
92 Endress+Hauser
Page 93
Proline Promass F 500
213 (8.4)
39 (1.5)
243 (9.6)
203 (8.0)
146 (5.75)
48 (1.9)
12 (0.47)
280 (11.0) 255 (10.0)
134 (5.3)
30 (1.18)
105 (4.1)
Protective cover
A0029552
 43 Weather protection cover for Proline 500 – digital
 44 Weather protection cover for Proline 500
External WLAN antenna
The external WLAN antenna is not suitable for use in hygienic applications.
Proline 500 – digital
External WLAN antenna mounted on device
 45 Engineering unit mm (in)
A0029553
A0033607
Endress+Hauser 93
Page 94
Proline Promass F 500
70 (2.8)
1500 (59.1)
72 (2.8)
68 (2.7)105 (4.1)
173 (6.8)
External WLAN antenna mounted with cable
The external WLAN antenna can be mounted separately from the transmitter if the transmission/ reception conditions at the transmitter mounting location are poor.
A0033606
 46 Engineering unit mm (in)
Proline 500
External WLAN antenna mounted on device
A0028923
 47 Engineering unit mm (in)
External WLAN antenna mounted with cable
The external WLAN antenna can be mounted separately from the transmitter if the transmission/ reception conditions at the transmitter mounting location are poor.
94 Endress+Hauser
Page 95
Proline Promass F 500
70 (2.8)
1500 (59.1)
72 (2.8)
A0033597
 48 Engineering unit mm (in)
Endress+Hauser 95
Page 96

Dimensions in US units Housing of Proline 500 – digital transmitter

N
P
F
G
A
Q
T
G
A
B
C
Q
Q
F
Non-hazardous area or hazardous area: Zone 2; Class I, Division 2
Order code for "Transmitter housing", option A "Aluminum, coated" and order code for "Integrated ISEM electronics", option A "Sensor"
Proline Promass F 500
A0033789
A
[in]
6.57 9.13 3.15 7.36 0.94 0.83
F
[in]
G
[in]
N
[in]
P
[in]
Q
[in]
Order code for "Transmitter housing", option D "Polycarbonate" and order code for "Integrated ISEM electronics", option A "Sensor"
A
[in]
6.97 9.21 3.54 7.76 0.67 0.87
F
[in]
G
[in]
N
[in]
P
[in]
Q
[in]
Housing of Proline 500 transmitter
Hazardous area: Zone 2; Class I, Division 2 or Zone 1; Class I, Division 1
Order code for "Transmitter housing", option A "Aluminum, coated" and order code for "Integrated ISEM electronics", option B "Transmitter"
A
[in]
7.40 3.35 4.06 12.5 8.54 5.12 9.41
B
[in]
C
[in]
F
[in]
G
[in]
Q
[in]
[in]
96 Endress+Hauser
A0033788
T
Page 97
Proline Promass F 500
A
B
C
G
D
L
E
F
K
M
Order code for "Transmitter housing", option L "Cast, stainless" and order code for "Integrated ISEM electronics", option B "Transmitter"
A
[in]
B
[in]
C
[in]
F
[in]
G
[in]
Q
[in]
7.40 3.35 4.06 11.6 8.54 5.12 9.41
Sensor with connection housing
Order code for "Sensor connection housing", option A "Aluminum, coated"
DN A
1)
[in] [in] [in] [in] [in] [in] [in] [in] [in] [in] [in]
³⁄₈ 5.83 3.70 2.13 2.95 7.28 10.24 5.35 0.211
½ 5.83 3.70 2.13 2.95 7.28 10.24 5.35 0.33
1 5.83 3.70 2.13 2.95 7.28 10.24 5.35 0.47
5.83 3.70 2.13 4.13 7.46 11.59 5.35 0.69
2 5.83 3.70 2.13 5.55 7.85 13.41 5.35 1.02
3 5.83 3.70 2.13 7.87 8.64 16.52 5.35 1.59
4 5.83 3.70 2.13 10 9.37 19.37 5.35 2.02
6 5.83 3.70 2.13 14.88 10.2 25.08 5.35 2.71
10 5.83 3.70 2.13 21.57 11.91 33.48 5.35 4.03
1)
B
C D E
2) 3)
F
2) 3)
G K L M
4)
4)
4)
4)
4)
4)
4)
4)
4)
T
[in]
A0033787
2.76
2.76
2.76
3.11
3.90
5.47
6.93
8.58
12.01
1) Depending on the cable gland used: values up to + 1.18 in
2) With order code for "Sensor option", option CG or order code for "Measuring tube material", option SD, SE, SF, TH, LA: values +2.76 in
3) With order code for "Measuring tube material", option TT, TU: values +4.09 in
4) Depends on the process connection in question →  99
Order code for "Sensor connection housing", option B "Stainless"
DN A
[in] [in] [in] [in] [in] [in] [in] [in] [in] [in] [in]
1)
³⁄₈ 5.39 3.07 2.32 2.95 7.09 10.04 5.28 0.211
½ 5.39 3.07 2.32 2.95 7.09 10.04 5.28 0.33
1 5.39 3.07 2.32 2.95 7.09 10.04 5.28 0.47
5.39 3.07 2.32 4.13 7.26 11.4 5.28 0.69
2 5.39 3.07 2.32 5.55 7.66 13.21 5.28 1.02
1)
B
C D E
2) 3)
F
2) 3)
G K L M
4)
4)
4)
4)
4)
2.76
2.76
2.76
3.11
3.90
Endress+Hauser 97
Page 98
Proline Promass F 500
DN A
1)
1)
B
C D E
2) 3)
F
2) 3)
G K L M
[in] [in] [in] [in] [in] [in] [in] [in] [in] [in] [in]
3 5.39 3.07 2.32 7.87 8.44 16.32 5.28 1.59
4 5.39 3.07 2.32 10 9.17 19.17 5.28 2.02
6 5.39 3.07 2.32 14.88 10 24.88 5.28 2.71
10 5.39 3.07 2.32 21.57 11.71 33.29 5.28 4.03
4)
4)
4)
4)
5.47
6.93
8.58
12.01
1) Depending on the cable gland used: values up to +1.18 in
2) With order code for "Sensor option", option CG or order code for "Measuring tube material", option SD, SE, SF, TH, LA: values +2.76 in
3) With order code for "Measuring tube material", option TT, TU: values +4.09 in
4) Depends on the process connection in question →  99
Order code for "Sensor connection housing", option C "Ultra-compact hygienic, stainless"
DN A
1)
[in] [in] [in] [in] [in] [in] [in] [in] [in] [in] [in]
³⁄₈ 4.88 2.68 2.20 2.95 7.09 10.04 4.41 0.21
½ 4.88 2.68 2.20 2.95 7.09 10.04 4.41 0.33
1 4.88 2.68 2.20 2.95 7.09 10.04 4.41 0.47
4.88 2.68 2.20 4.13 7.26 11.4 4.41 0.69
2 4.88 2.68 2.20 5.55 7.66 13.21 4.41 1.02
3 4.88 2.68 2.20 7.87 8.44 16.32 4.41 1.59
4 4.88 2.68 2.20 10 9.17 19.17 4.41 2.02
6 4.88 2.68 2.20 14.88 10 24.88 4.41 2.71
10 4.88 2.68 2.20 21.57 11.71 33.29 4.41 4.03
1)
B
C D E
2) 3)
F
2) 3)
G K L M
4)
4)
4)
4)
4)
4)
4)
4)
4)
2.76
2.76
2.76
3.11
3.90
5.47
6.93
8.58
12.01
1) Depending on the cable gland used: values up to +1.18 in
2) With order code for "Sensor option", option CG or order code for "Measuring tube material", option SD, SE, SF, TH, LA: values +2.76 in
3) With order code for "Measuring tube material", option TT, TU: values +4.09 in
4) Depends on the process connection in question →  99
Order code for "Sensor connection housing", option L "Cast, stainless"
DN A
1)
[in] [in] [in] [in] [in] [in] [in] [in] [in] [in] [in]
³⁄₈ 5.71 3.39 2.32 2.95 8.19 11.14 5.35 0.211
½ 5.71 3.39 2.32 2.95 8.19 11.14 5.35 0.33
1 5.71 3.39 2.32 2.95 8.19 11.14 5.35 0.47
5.71 3.39 2.32 4.13 8.37 12.5 5.35 0.69
2 5.71 3.39 2.32 5.55 8.76 14.31 5.35 1.02
3 5.71 3.39 2.32 7.87 9.55 17.42 5.35 1.59
4 5.71 3.39 2.32 10 10.28 20.28 5.35 2.02
6 5.71 3.39 2.32 14.88 11.1 25.98 5.35 2.71
10 5.71 3.39 2.32 21.57 12.81 34.39 5.35 4.03
1) Depending on the cable gland used: values up to +1.18 in
2) With order code for "Sensor option", option CG or order code for "Measuring tube material", option SD, SE, SF, TH, LA: values +2.76 in
3) With order code for "Measuring tube material", option TT, TU: values +4.09 in
4) Depends on the process connection in question →  99
1)
B
C D E
2) 3)
F
2) 3)
G K L M
4)
4)
4)
4)
4)
4)
4)
4)
4)
2.76
2.76
2.76
3.11
3.90
5.47
6.93
8.58
12.01
98 Endress+Hauser
Page 99
Proline Promass F 500
C
D
A
B
E
L
Flange connections
Fixed flange ASME B16.5
A0015621
Length tolerance for dimension L in inch:
• DN ≤ 4": +0.06 / –0.08
• DN ≥ 5": +0.14
Flange according to ASME B16.5: Class 150
1.4404 (F316/F316L): order code for "Process connection", option AAS Alloy C22: order code for "Process connection", option AAC
DN
[in]
1)
³⁄₈
½ 3.54 2.37 4 × ⌀0.62 0.44 0.62 15.91
1 4.33 3.13 4 × ⌀0.62 0.56 1.05 17.32
4.92 3.87 4 × ⌀0.62 0.69 1.61 21.65
2 5.91 4.75 4 × ⌀0.75 0.75 2.07 28.15
3 7.48 6.00 4 × ⌀0.75 0.94 3.07 33.07
4 9.06 7.50 8 × ⌀0.75 0.94 4.03 44.37
6 11.02 9.50 8 × ⌀0.88 1 6.07 55.04
10 15.94 14.25 12 × ⌀1.0 1.19 10.02 72.13
Surface roughness (flange): Ra 125 to 250 µin
1) DN ³⁄₈" with DN ½" flanges as standard
Flange according to ASME B16.5: Class 150 with reduction in nominal diameter
1.4404 (F316/F316L)
DN
[in]
2 AHS 5.91 4.75 4 × ⌀ 0.75 0.75 2.07 21.65
3 2 AJS 7.48 6 4 × ⌀ 0.75 0.94 3.07 28.35
4 3 ALS 9.06 7.5 8 × ⌀ 0.75 0.94 4.03 34.41
6 4 ANS 11.02 9.5 8 × ⌀ 0.88 1 6.07 45.94
8 6 APS 13.58 11.75 8 × ⌀ 0.88 1.14 7.98 57.52
Surface roughness (flange): Ra 125 to 250 µin
A
[in]
3.54 2.37 4 × ⌀ 0.62 0.44 0.62 14.57
reduction
to DN
[in]
Order code for
B
[in]
"Process
connection",
option
A
[in]
C
[in]
B
[in]
D
[in]
C
[in]
E
[in]
D
[in]
E
[in]
L
[in]
L
[in]
Endress+Hauser 99
Page 100
Proline Promass F 500
Flange according to ASME B16.5: Class 300
1.4404 (F316/F316L): order code for "Process connection", option ABS Alloy C22: order code for "Process connection", option ABC
DN
[in]
1)
³⁄₈
½ 3.74 2.63 4 × ⌀0.62 0.56 0.62 15.91
1 4.92 3.50 4 × ⌀0.75 0.69 1.05 17.32
6.10 4.50 4 × ⌀0.88 0.81 1.61 21.65
2 6.50 5.00 8 × ⌀0.75 0.88 2.07 28.15
3 8.27 6.63 8 × ⌀0.88 1.12 3.07 33.07
4 10.04 7.87 8 × ⌀0.88 1.25 4.03 44.37
6 12.6 10.63 12 × ⌀0.88 1.44 6.07 55.79
10 17.52 15.25 16 × ⌀1.12 1.87 10.02 73.35
Surface roughness (flange): Ra 125 to 250 µin
1) DN ³⁄₈" with DN ½" flanges as standard
A
[in]
B
[in]
C
[in]
D
[in]
E
[in]
3.74 2.63 4 × ⌀0.62 0.56 0.62 14.57
L
[in]
Flange according to ASME B16.5: Class 300 with reduction in nominal diameter
1.4404 (F316/F316L)
DN
[in]
reduction
to DN
[in]
Order code for
"Process
connection",
A
[in]
B
[in]
C
[in]
D
[in]
E
[in]
option
2 AIS 6.5 5 8 × ⌀ 0.75 0.88 2.07 24.21
3 2 AKS 8.27 6.63 8 × ⌀ 0.88 1.12 3.07 28.82
4 3 AMS 10.04 7.87 8 × ⌀ 0.88 1.25 4.03 35.2
6 4 AOS 12.6 10.63 12 × ⌀ 0.88 1.44 6.07 46.73
8 6 AQS 14.96 13 12 × ⌀ 1 1.64 7.98 57.52
Surface roughness (flange): Ra 125 to 250 µin
Flange according to ASME B16.5: Class 600
1.4404 (F316/F316L): order code for "Process connection", option ACS Alloy C22: order code for "Process connection", option ACC
DN
[in]
³⁄₈
1)
A
[in]
B
[in]
C
[in]
D
[in]
E
[in]
3.74 2.63 4 × ⌀0.62 0.81 0.55 15.75
½ 3.74 2.63 4 × ⌀0.62 0.81 0.55 16.54
1 4.92 3.50 4 × ⌀0.75 0.94 0.96 19.29
6.10 4.50 4 × ⌀0.88 1.13 1.5 23.62
2 6.50 5.00 8 × ⌀0.75 1.25 1.94 29.21
3 8.27 6.63 8 × ⌀0.88 1.5 2.9 35.43
4 10.83 8.50 8 × ⌀1.00 1.91 3.83 45.55
6 13.98 11.50 12 × ⌀1.12 1.88 6.07 57.76
10 20.08 17.00 16 × ⌀1.38 2.75 10.02 76.61
Surface roughness (flange): Ra 125 to 250 µin
L
[in]
L
[in]
1) DN ³⁄₈" with DN ½" flanges as standard
100 Endress+Hauser
Loading...