Endress+Hauser F 500 Specifications

TI01222D/06/EN/04.18 71402584 2018-06-01
Products Solutions Services
Technical Information
Proline Promass F 500
Coriolis flowmeter
The flowmeter offering premium accuracy and robustness, available as remote version with up to 4 I/Os
Application
• Measuring principle operates independently of physical fluid properties such as viscosity or density
• Highest measurement performance for liquids and gases under varying, demanding process conditions
Device properties
• Medium temperature: –196 to +350 °C (–320 to +662 °F)
• Nominal diameter: DN 8 to 250 (³⁄₈ to 10")
• Remote version with up to 4 I/Os
• Backlit display with touch control and WLAN access
• Standard cable between sensor and transmitter
Your benefits
• Highest process safety – immune to fluctuating and harsh environments
• Fewer process measuring points – multivariable measurement (flow, density, temperature)
• Spacesaving installation – no inlet/outlet run needs
• Full access to process and diagnostic information – numerous, freely combinable I/Os and fieldbuses
• Reduced complexity and variety – freely configurable I/O functionality
• Integrated verification – Heartbeat Technology

Table of contents

Proline Promass F 500
About this document ........................ 4
Symbols used ................................ 4
Function and system design ................... 5
Measuring principle ............................ 5
Measuring system ............................. 7
Equipment architecture ......................... 9
Safety ..................................... 9
Input .................................... 12
Measured variable ............................ 12
Measuring range ............................. 12
Operable flow range ........................... 13
Input signal ................................ 13
Output .................................. 15
Output and input variants ....................... 15
Output signal ............................... 16
Signal on alarm .............................. 20
Ex connection data ........................... 22
Low flow cut off ............................. 23
Galvanic isolation ............................ 23
Protocol-specific data .......................... 24
Power supply ............................. 28
Terminal assignment .......................... 28
Device plugs available .......................... 29
Pin assignment, device plug ...................... 30
Supply voltage .............................. 32
Power consumption ........................... 32
Current consumption .......................... 32
Power supply failure .......................... 32
Electrical connection .......................... 33
Potential equalization ......................... 43
terminals .................................. 44
Cable entries ............................... 44
Cable specification ............................ 44
Performance characteristics .................. 49
Reference operating conditions ................... 49
Maximum measured error ....................... 49
Repeatability ............................... 51
Response time .............................. 52
Influence of ambient temperature ................. 52
Influence of medium temperature .................. 52
Influence of medium pressure .................... 53
Design fundamentals .......................... 53
Installation ............................... 54
Mounting location ............................ 54
Orientation ................................ 55
Inlet and outlet runs .......................... 56
Mounting the transmitter housing ................. 57
Special mounting instructions .................... 58
Environment .............................. 60
Ambient temperature range ..................... 60
Storage temperature .......................... 60
Climate class ............................... 60
Degree of protection .......................... 60
Vibration resistance ........................... 60
Shock resistance ............................. 61
Shock resistance ............................. 61
Interior cleaning ............................. 61
Electromagnetic compatibility (EMC) ............... 61
Process .................................. 61
Medium temperature range ...................... 61
Density ................................... 62
Pressure-temperature curves ..................... 63
Sensor housing .............................. 68
Rupture disk ................................ 69
Flow limit ................................. 69
Pressure loss ............................... 69
System pressure ............................. 69
Thermal insulation ........................... 70
Heating ................................... 70
Vibrations ................................. 71
Custody transfer measurement ............... 72
Mechanical construction .................... 73
Dimensions in SI units ......................... 73
Dimensions in US units ......................... 96
Weight .................................. 108
Materials ................................. 109
Process connections .......................... 112
Surface roughness ........................... 112
Operability .............................. 112
Operating concept ........................... 112
Languages ................................ 113
Local operation ............................. 113
Remote operation ........................... 113
Service interface ............................ 119
Network integration .......................... 120
Supported operating tools ...................... 121
HistoROM data management .................... 122
Certificates and approvals .................. 124
CE mark .................................. 124
C-Tick symbol .............................. 124
Ex approval ............................... 124
Sanitary compatibility ......................... 127
Pharmaceutical compatibility .................... 127
Functional safety ............................ 127
HART certification ........................... 127
FOUNDATION Fieldbus certification ............... 127
Certification PROFIBUS ........................ 127
EtherNet/IP certification ....................... 128
Certification PROFINET ........................ 128
Pressure Equipment Directive ................... 128
Radio approval ............................. 128
Measuring instrument approval .................. 128
Additional certification ........................ 128
2 Endress+Hauser
Proline Promass F 500
Other standards and guidelines .................. 129
Ordering information ...................... 130
Application packages ...................... 130
Diagnostics functions ......................... 130
Heartbeat Technology ........................ 130
Concentration .............................. 131
Special density ............................. 131
Petroleum ................................ 131
OPC-UA server ............................. 131
Accessories .............................. 131
Device-specific accessories ...................... 132
Communication-specific accessories ............... 133
Service-specific accessories ..................... 134
System components .......................... 134
Supplementary documentation .............. 135
Standard documentation ....................... 135
Device-dependent additional documentation ......... 136
Registered trademarks ..................... 136
Endress+Hauser 3

About this document

A

Symbols used Electrical symbols

Symbol Meaning
Proline Promass F 500
Direct current
Alternating current
Direct current and alternating current
Ground connection
A grounded terminal which, as far as the operator is concerned, is grounded via a grounding system.
Protective Earth (PE)
A terminal which must be connected to ground prior to establishing any other connections.
The ground terminals are situated inside and outside the device:
• Inner ground terminal: Connects the protectiv earth to the mains supply.
• Outer ground terminal: Connects the device to the plant grounding system.
Communication symbols
Symbol Meaning
Wireless Local Area Network (WLAN)
Communication via a wireless, local network.
LED
Light emitting diode is off.
LED
Light emitting diode is on.
LED
Light emitting diode is flashing.
Symbols for certain types of information
Symbol Meaning
Permitted
Procedures, processes or actions that are permitted.
Preferred
Procedures, processes or actions that are preferred.
Forbidden
Procedures, processes or actions that are forbidden.
Tip
Indicates additional information.
Reference to documentation.
Reference to page.
Reference to graphic.
Visual inspection.
4 Endress+Hauser
Proline Promass F 500
1.
-
.
21 3
Symbols in graphics
Symbol Meaning
1, 2, 3, ... Item numbers
, 2., 3., … Series of steps
A, B, C, ... Views
A-A, B-B, C-C, ... Sections
Hazardous area
Safe area (non-hazardous area)
Flow direction

Function and system design

Measuring principle

The measuring principle is based on the controlled generation of Coriolis forces. These forces are always present in a system when both translational and rotational movements are superimposed.
Fc = 2 · ∆m (ν · ω)
Fc = Coriolis force
∆m = moving mass
ω = rotational velocity
ν = radial velocity in rotating or oscillating system
The amplitude of the Coriolis force depends on the moving mass ∆m, its velocity ν in the system and thus on the mass flow. Instead of a constant rotational velocity ω, the sensor uses oscillation.
In the sensor, two parallel measuring tubes containing flowing fluid oscillate in antiphase, acting like a tuning fork. The Coriolis forces produced at the measuring tubes cause a phase shift in the tube oscillations (see illustration):
• At zero flow (when the fluid is at a standstill) the two tubes oscillate in phase (1).
• Mass flow causes deceleration of the oscillation at the inlet of the tubes (2) and acceleration at the outlet (3).
A0028850
The phase difference (A-B) increases with increasing mass flow. Electrodynamic sensors register the tube oscillations at the inlet and outlet. System balance is ensured by the antiphase oscillation of the two measuring tubes. The measuring principle operates independently of temperature, pressure, viscosity, conductivity and flow profile.
Endress+Hauser 5
Proline Promass F 500
Density measurement
The measuring tube is continuously excited at its resonance frequency. A change in the mass and thus the density of the oscillating system (comprising measuring tube and fluid) results in a corresponding, automatic adjustment in the oscillation frequency. Resonance frequency is thus a function of medium density. The microprocessor utilizes this relationship to obtain a density signal.
Volume measurement
Together with the measured mass flow, this is used to calculate the volume flow.
Temperature measurement
The temperature of the measuring tube is determined in order to calculate the compensation factor due to temperature effects. This signal corresponds to the process temperature and is also available as an output signal.
6 Endress+Hauser
Proline Promass F 500
2
3
1
A
B
1
2
3

Measuring system

The measuring system consists of a transmitter and a sensor. The transmitter and sensor are mounted in physically separate locations. They are interconnected by connecting cables.
Transmitter
Two versions of the transmitter are available.
Proline 500 – digital Proline 500
For use in applications not required to meet special requirements due to ambient or operating conditions.
A Non-hazardous area or Zone 2; Class I, Division 2 B Non-hazardous area or Zone 2; Class I, Division 2 or Zone 1; Class I,
Division 1 1 Transmitter 2 Connecting cable: cable, separate, standard 3 Sensor connection housing with integrated ISEM
• Flexible and cost-effective separate installation.
• A standard cable can be used as the connecting cable.
• Electronics in the transmitter housing, ISEM (intelligent sensor
electronics module) in the sensor connection housing
• Signal transmission: digital
Order code for "Integrated ISEM electronics", option A "Sensor"
Connecting cable (can be ordered in various lengths)→  132
• Length:
– Zone 2; Class I, Division 2: max. 300 m (1 000 ft) – Zone 1; Class I, Division 1: max. 150 m (500 ft)
• Standard cable with common shield (pair-stranded)
Hazardous area
Use in: Zone 2; Class I, Division 2
Mixed installation is possible:
• Sensor: Zone 1; Class I, Division 1
• Transmitter: Zone 2; Class I, Division 2
Device versions and materials
• Transmitter housing
– Aluminum, coated: aluminum, AlSi10Mg, coated – Material: polycarbonate
• Material of window in transmitter housing
– Aluminum, coated: glass – Polycarbonate: plastic
Configuration
• External operation via 4-line, backlit, graphic local display with touch control and guided menus ("Make-it-run" wizards) for application-specific
commissioning.
• Via service interface or WLAN interface:
– Operating tools (e.g. FieldCare, DeviceCare, SmartBlue app) – Web server (access via Web browser, e.g. Microsoft Internet Explorer, Microsoft Edge)
For use in applications required to meet special requirements due to ambient or operating conditions.
Non-hazardous area or Zone 2; Class I, Division 2 or Zone 1; Class I,
Division 1 1 Transmitter with integrated ISEM 2 Connecting cable: cable, separate 3 Sensor connection housing
Application examples for sensors without electronics:
• Strong vibrations at the sensor.
• Sensor in underground installations.
• Permanent immersion of sensor in water, IP68 ingress protection.
• Electronics and ISEM (intelligent sensor electronics module) in the
transmitter housing
• Signal transmission: analog
Order code for "Integrated ISEM electronics", option B "Transmitter"
• Length: max. 20 m (65 ft)
• Cable with a common shield and individual shielded cores (3 pairs)
Use in: Zone 1; Class I, Division 1 oder Zone 2; Class I, Division 2
• Transmitter housing
– Aluminum, coated: aluminum, AlSi10Mg, coated – Cast, stainless: cast, stainless steel, 1.4409 (CF3M) similar to 316L
• Window material: glass
Endress+Hauser 7
Sensor connection housing
Different versions of the connection housing are available.
Order code for "Sensor connection housing", option A, "Aluminum, coated": Aluminum, AlSi10Mg, coated
This device version is only available in conjunction with the Proline
500 – digital transmitter.
Order code for "Sensor connection housing", option B, "Stainless":
• Hygienic version, stainless steel 1.4301 (304)
• Optional: order code for "Sensor feature", option CC "Hygienic version, for maximum corrosion resistance": stainless steel 1.4404 (316L)
Order code for "Sensor connection housing", option C, "Ultra-compact hygienic, stainless":
• Hygienic version, stainless steel 1.4301 (304)
• Optional: order code for "Sensor feature", option CC "Hygienic version, for maximum corrosion resistance": stainless steel 1.4404 (316L)
This device version is only available in conjunction with the Proline
500 – digital transmitter.
Order code for "Sensor connection housing", option L, "Cast, stainless":
1.4409 (CF3M) similar to 316L
Proline Promass F 500
Sensor
PromassF • Excellent performance across a wide range of applications
• Simultaneous measurement of flow, volume flow, density and temperature (multivariable)
• Immune to process influences
• Materials:
A0026713
– Sensor: stainless steel, 1.4301/1.4307 (304L); optional 1.4404
(316/316L)
– Measuring tubes: stainless steel, 1.4539 (904L); 1.4404
(316/316L); Alloy C22, 2.4602 (UNS N06022)
– Process connections: stainless steel, 1.4404 (316/316L); 1.4301
(304); Alloy C22, 2.4602 (UNS N06022)
8 Endress+Hauser
Proline Promass F 500
2
1
6
5
7
4
3
3

Equipment architecture

A0027512
 1 Possibilities for integrating measuring devices into a system
1 Control system (e.g. PLC) 2 Connecting cable (0/4 to 20 mA HART etc.) 3 Fieldbus 4 Segment coupler 5 Non-hazardous area 6 Hazardous area: Zone 2; Class I, Division 2 7 Hazardous area: Zone 1; Class I, Division 1

Safety IT security

Our warranty is valid only if the device is installed and used as described in the Operating Instructions. The device is equipped with security mechanisms to protect it against any inadvertent changes to the settings.
IT security measures, which provide additional protection for the device and associated data transfer, must be implemented by the operators themselves in line with their security standards.
Device-specific IT security
The device offers a range of specific functions to support protective measures on the operator's side. These functions can be configured by the user and guarantee greater in-operation safety if used correctly. An overview of the most important functions is provided in the following section.
Function/interface Factory setting Recommendation
Write protection via hardware write protection switch →  10
Access code (also applies for Web server login or FieldCare connection) →  10
WLAN (order option in display module)
WLAN security mode Enabled (WPA2-
WLAN passphrase (password) →  10
WLAN mode Access Point On an individual basis following risk
Web server→  10 Enabled. On an individual basis following risk
CDI-RJ45 service interface →  11 On an individual basis following risk
Not enabled. On an individual basis following risk
assessment.
Not enabled (0000).
Enabled. On an individual basis following risk
PSK)
Serial number Assign a customized access code during
Assign a customized access code during commissioning.
assessment.
Do not change.
commissioning.
assessment.
assessment.
assessment.
Endress+Hauser 9
Proline Promass F 500
Protecting access via hardware write protection
Write access to the device parameters via the local display, Web browser or operating tool (e.g. FieldCare, DeviceCare) can be disabled via a write protection switch (DIP switch on the motherboard). When hardware write protection is enabled, only read access to the parameters is possible.
Hardware write protection is disabled when the device is delivered.
Protecting access via a password
Different passwords are available to protect write access to the device parameters or access to the device via the WLAN interface.
• User-specific access code Protect write access to the device parameters via the local display, Web browser or operating tool (e.g. FieldCare, DeviceCare). Access authorization is clearly regulated through the use of a user­specific access code.
• WLAN passphrase The network key protects a connection between an operating unit (e.g. notebook or tablet) and the device via the WLAN interface which can be ordered as an option.
• Infrastructure mode When the device is operated in infrastructure mode, the WLAN passphrase corresponds to the WLAN passphrase configured on the operator side.
User-specific access code
Write access to the device parameters via the local display, Web browser or operating tool (e.g. FieldCare, DeviceCare) can be protected by the modifiable, user-specific access code.
WLAN passphrase: Operation as WLAN access point
A connection between an operating unit (e.g. notebook or tablet) and the device via the WLAN interface, which can be ordered as an optional extra, is protected by the network key. The WLAN authentication of the network key complies with the IEEE 802.11 standard.
When the device is delivered, the network key is pre-defined depending on the device. It can be changed via the WLAN settings submenu in the WLAN passphrase parameter.
Infrastructure mode
A connection between the device and WLAN access point is protected by means of an SSID and passphrase on the system side. Please contact the relevant system administrator for access.
General notes on the use of passwords
• The access code and network key supplied with the device should be changed during commissioning.
• Follow the general rules for generating a secure password when defining and managing the access code or network key.
• The user is responsible for the management and careful handling of the access code and network key.
Access via Web server
The device can be operated and configured via a Web browser with the integrated Web server. The connection is via the service interface (CDI-RJ45) or the WLAN interface. For device versions with the EtherNet/IP and PROFINET communication protocols, the connection can also be established via the terminal connection for signal transmission with EtherNet/IP or PROFINET (RJ45 connector).
The Web server is enabled when the device is delivered. The Web server can be disabled if necessary (e.g. after commissioning) via the Web server functionality parameter.
The device and status information can be hidden on the login page. This prevents unauthorized access to the information.
For detailed information on device parameters, see: The "Description of Device Parameters" document →  135
Access via OPC-UA
The "OPC UA Server" application package is available in the device version with the HART communication protocol →  131.
10 Endress+Hauser
Proline Promass F 500
The device can communicate with OPC UA clients using the "OPC UA Server" application package.
The OPC UA server integrated in the device can be accessed via the WLAN access point using the WLAN interface - which can be ordered as an optional extra - or the service interface (CDI- RJ45) via Ethernet network. Access rights and authorization as per separate configuration.
The following Security Modes are supported as per the OPC UA Specification (IEC 62541):
• None
• Basic128Rsa15 – signed
• Basic128Rsa15 – signed and encrypted
Access via service interface (CDI-RJ45)
The device can be connected to a network via the service interface (CDI-RJ45). Device-specific functions guarantee the secure operation of the device in a network.
The use of relevant industrial standards and guidelines that have been defined by national and international safety committees, such as IEC/ISA62443 or the IEEE, is recommended. This includes organizational security measures such as the assignment of access authorization as well as technical measures such as network segmentation.
The device can be integrated in a ring topology. The device is integrated via the terminal connection for signal transmission (output 1) and the connection to the service interface (CDI­RJ45) →  117.
Endress+Hauser 11

Input

Measured variable Direct measured variables

• Mass flow
• Density
• Temperature
Calculated measured variables
• Volume flow
• Corrected volume flow
• Reference density

Measuring range Measuring ranges for liquids

Proline Promass F 500
DN Measuring range full scale values 
min(F)
to 
[mm] [in] [kg/h] [lb/min]
8 ³⁄₈ 0 to 2 000 0 to 73.50
15 ½ 0 to 6 500 0 to 238.9
25 1 0 to 18 000 0 to 661.5
40 0 to 45 000 0 to 1 654
50 2 0 to 70 000 0 to 2 573
80 3 0 to 180 000 0 to 6 615
100 4 0 to 350 000 0 to 12 860
150 6 0 to 800 000 0 to 29 400
250 10 0 to 2 200 000 0 to 80 850
Measuring range for gases
The full scale value depends on the density and the sound velocity of the gas used and can be calculated with the formula below: 
= minimum (
max(G)
max(G)
max(F)
< 
ρ
max(G)
G
max(F)
x Constant dependent on nominal diameter
c
G
d
i
· ρG : x ; ρG · cG · π/2 · (di)2 · 3600)
max(F)
Maximum full scale value for gas [kg/h]
Maximum full scale value for liquid [kg/h]
can never be greater than 
max(G)
Gas density in [kg/m³] at operating conditions
Sound velocity (gas) [m/s]
Measuring tube internal diameter [m]
max(F)
max(F)
DN x
[mm] [in] [kg/m3]
8 ³⁄₈ 60
15 ½ 80
25 1 90
40 90
50 2 90
80 3 110
12 Endress+Hauser
Proline Promass F 500
DN x
[mm] [in] [kg/m3]
100 4 130
150 6 200
250 10 200
To calculate the measuring range, use the Applicator sizing tool →  134
Calculation example for gas
• Sensor: Promass F, DN 50
• Gas: Air with a density of 60.3 kg/m³ (at 20 °C and 50 bar)
• Measuring range (liquid): 70 000 kg/h
• x = 90 kg/m³ (for Promass F, DN 50) Maximum possible full scale value:
max(G)
= 
· ρG : x = 70 000 kg/h · 60.3 kg/m³ : 90 kg/m³ = 46 900 kg/h
max(F)
Recommended measuring range
"Flow limit" section →  69

Operable flow range

Over 1000 : 1.
Flow rates above the preset full scale value do not override the electronics unit, with the result that the totalizer values are registered correctly.

Input signal Input and output versions

→  15
External measured values
To increase the accuracy of certain measured variables or to calculate the corrected volume flow for gases, the automation system can continuously write different measured values to the measuring device:
• Operating pressure to increase accuracy (Endress+Hauser recommends the use of a pressure measuring device for absolute pressure, e.g. Cerabar M or Cerabar S)
• Medium temperature to increase accuracy (e.g. iTEMP)
• Reference density for calculating the corrected volume flow for gases
Various pressure transmitters and temperature measuring devices can be ordered from Endress +Hauser: see "Accessories" section →  134
It is recommended to read in external measured values to calculate the corrected volume flow.
HART protocol
The measured values are written from the automation system to the measuring device via the HART protocol. The pressure transmitter must support the following protocol-specific functions:
• HART protocol
• Burst mode
Current input
The measured values are written from the automation system to the measuring device via the current input →  14.
Digital communication
The measured values can be written from the automation system to the measuring via:
• FOUNDATION Fieldbus
• PROFIBUS DP
• PROFIBUS PA
• Modbus RS485
• EtherNet/IP
• PROFINET
Endress+Hauser 13
Current input 0/4 to 20 mA
Current input 0/4 to 20 mA (active/passive)
Current span • 4 to 20 mA (active)
• 0/4 to 20 mA (passive)
Resolution 1 µA
Voltage drop Typically: 0.6 to 2 V for 3.6 to 22 mA (passive)
Maximum input voltage ≤ 30 V (passive)
Open-circuit voltage ≤ 28.8 V (active)
Possible input variables • Pressure
• Temperature
• Density
Status input
Maximum input values • DC –3 to 30 V
• If status input is active (ON): Ri >3 kΩ
Response time Adjustable: 5 to 200 ms
Input signal level • Low signal: DC –3 to +5 V
• High signal: DC 12 to 30 V
Assignable functions • Off
• Reset the individual totalizers separately
• Reset all totalizers
• Flow override
Proline Promass F 500
14 Endress+Hauser
Proline Promass F 500

Output

Output and input variants

Depending on the option selected for output/input 1, different options are available for the other outputs and inputs. Only one option can be selected for each output/input 1 to 4. The table must be read vertically (↓).
Example: If the option BA "4–20 mA HART" was selected for output/input 1, one of the options A, B, D, E, F, H, I or J is available for output 2, and one of the options A, B, D, E, F, H, I or J is available for output 3 and 4.
Order code for "Output; input 1" (020) → Possible options
Current output 4 to 20 mA HART BA
Current output 4 to 20 mA HART Ex i CA
FOUNDATION Fieldbus SA
FOUNDATION Fieldbus Ex i TA
PROFIBUS DP LA
PROFIBUS PA GA
PROFIBUS PA Ex i HA
Modbus RS485 MA
EtherNet/IP 2-port switch integrated NA
PROFINET 2-port switch integrated RA
Order code for "Output; input 2" (021) →
Not assigned A A A A A A A A A A
Current output 0/4 to 20 mA B B B B B B B
Current output 0/4 to 20 mA (Ex i) C C C
User configurable input/output
Pulse/frequency/switch output E E E E E E E
Double pulse output
Pulse/frequency/switch output (Ex i) G G G
Relay output H H H H H H H
Current input 0/4 to 20 mA I I I I I I I
Status input J J J J J J J
Order code for "Output; input 3" (022), "Output; input 4" (023)
Not assigned A A A A A A A A A A
Current output 0/4 to 20 mA B B B B B
Current output 0/4 to 20 mA (Ex i) C
User configurable input/output D D D D D
Pulse/frequency/switch output E E E E E
Double pulse output (slave)
Pulse/frequency/switch output (Ex i) G
Relay output H H H H H
Current input 0/4 to 20 mA I I I I I
Status input J J J J J
2)
1)
3)
4)
D D D D D D D
F F
F F
1) A specific input or output can be assigned to a user configurable input/output →  19.
2) If double pulse output (F) is selected for output/input 2 (021), only the double pulse output (F) option is available for selection for output/input 3 (022).
3) The order code for "Output; input 4" (023) is only available for the Proline 500 – digital transmitter.
4) The double pulse output (F) option is not available for input/output 4.
Endress+Hauser 15

Output signal HART current output

Current output 4 to 20 mA HART
Current span Can be set to: 4 to 20 mA (active/passive)
Open-circuit voltage DC 28.8 V (active)
Maximum input voltage DC 30 V (passive)
Load 250 to 700 Ω
Resolution 0.38 µA
Damping Configurable: 0.07 to 999 s
Assignable measured variables
Proline Promass F 500
Ex-i, passive
• Mass flow
• Volume flow
• Corrected volume flow
• Density
• Reference density
• Temperature
• Electronic temperature
• Oscillation frequency 0
• Oscillation damping 0
• Signal asymmetry
• Exciter current 0
The range of options increases if the measuring device has one or more
application packages.
PROFIBUS PA
PROFIBUS PA In accordance with EN 50170 Volume 2, IEC 61158-2 (MBP), galvanically
isolated
Data transmission 31.25 kbit/s
Current consumption 10 mA
16 mA
Permitted supply voltage 9 to 32 V
Bus connection With integrated reverse polarity protection
PROFIBUS DP
Signal encoding NRZ code
Data transfer 9.6 kBaud…12 MBaud
EtherNet/IP
Standards In accordance with IEEE 802.3
PROFINET
Standards In accordance with IEEE 802.3
FOUNDATION Fieldbus
FOUNDATION Fieldbus H1, IEC 61158-2, galvanically isolated
Data transfer 31.25 kbit/s
16 Endress+Hauser
Proline Promass F 500
Current consumption 10 mA
Permitted supply voltage 9 to 32 V
Bus connection With integrated reverse polarity protection
Modbus RS485
Physical interface RS485 in accordance with EIA/TIA-485 standard
Terminating resistor Integrated, can be activated via DIP switches
Current output 0/4 to 20 mA
Current output 0/4 to 20 mA
Maximum output values 22.5 mA
Current span Can be set to:
• 4 to 20 mA (active)
• 0/4 to 20 mA (passive)
Ex-i, passive
Open-circuit voltage DC 28.8 V (active)
Maximum input voltage DC 30 V (passive)
Load 0 to 700 Ω
Resolution 0.38 µA
Damping Adjustable: 0.07 to 999 s
Assignable measured variables
• Mass flow
• Volume flow
• Corrected volume flow
• Density
• Reference density
• Temperature
• Electronic temperature
• Oscillation frequency 0
• Oscillation damping 0
• Signal asymmetry
• Exciter current 0
The range of options increases if the measuring device has one or more
application packages.
Pulse/frequency/switch output
Function Can be set to pulse, frequency or switch output
Version Open collector
Can be set to:
• Active
• Passive
Ex-i, passive
Maximum input values DC 30 V, 250 mA (passive)
Open-circuit voltage DC 28.8 V (active)
Voltage drop For 22.5 mA: ≤ DC 2 V
Pulse output
Maximum input values DC 30 V, 250 mA (passive)
Maximum output current 22.5 mA (active)
Endress+Hauser 17
Open-circuit voltage DC 28.8 V (active)
Pulse width Adjustable: 0.05 to 2 000 ms
Maximum pulse rate 10 000 Impulse/s
Pulse value Adjustable
Assignable measured variables
Frequency output
Maximum input values DC 30 V, 250 mA (passive)
Maximum output current 22.5 mA (active)
Open-circuit voltage DC 28.8 V (active)
Output frequency Adjustable: end value frequency 2 to 10 000 Hz (f
Damping Adjustable: 0 to 999 s
Pulse/pause ratio 1:1
Assignable measured variables
Switch output
Maximum input values DC 30 V, 250 mA (passive)
Open-circuit voltage DC 28.8 V (active)
Switching behavior Binary, conductive or non-conductive
Switching delay Adjustable: 0 to 100 s
Number of switching cycles
Assignable functions • Off
• Mass flow
• Volume flow
• Corrected volume flow
• Mass flow
• Volume flow
• Corrected volume flow
• Density
• Reference density
• Temperature
• Electronic temperature
• Oscillation frequency 0
• Oscillation damping 0
• Signal asymmetry
• Exciter current 0
The range of options increases if the measuring device has one or more
application packages.
Unlimited
• On
• Diagnostic behavior
• Limit value – Mass flow – Volume flow – Corrected volume flow – Density – Reference density – Temperature – Totalizer 1-3
• Flow direction monitoring
• Status – Partially filled pipe detection – Low flow
The range of options increases if the measuring device has one or more
application packages.
Proline Promass F 500
= 12 500 Hz)
max
18 Endress+Hauser
Proline Promass F 500
Double pulse output
Function Double pulse
Version Open collector
Can be set to:
• Active
• Passive
Maximum input values DC 30 V, 250 mA (passive)
Open-circuit voltage DC 28.8 V (active)
Voltage drop For 22.5 mA: ≤ DC 2 V
Output frequency Adjustable: 0 to 1 000 Hz
Damping Adjustable: 0 to 999 s
Pulse/pause ratio 1:1
Assignable measured variables
• Mass flow
• Volume flow
• Corrected volume flow
• Density
• Reference density
• Temperature
The range of options increases if the measuring device has one or more
application packages.
Relay output
Function Switch output
Version Relay output, galvanically isolated
Switching behavior Can be set to:
• NO (normally open), factory setting
• NC (normally closed)
Maximum switching capacity (passive)
Assignable functions • Off
• DC 30 V, 0.1 A
• AC 30 V, 0.5 A
• On
• Diagnostic behavior
• Limit value – Mass flow – Volume flow – Corrected volume flow – Density – Reference density – Temperature – Totalizer 1-3
• Flow direction monitoring
• Status – Partially filled pipe detection – Low flow
The range of options increases if the measuring device has one or more
application packages.
User configurable input/output
One specific input or output is assigned to a user-configurable input/output (configurable I/O)
during device commissioning.
The following inputs and outputs are available for assignment:
• Choice of current output: 4 to 20 mA (active), 0/4 to 20 mA (passive)
• Pulse/frequency/switch output
• Choice of current input: 4 to 20 mA (active), 0/4 to 20 mA (passive)
• Status input
Endress+Hauser 19
Proline Promass F 500
The technical values correspond to those of the inputs and outputs described in this section.

Signal on alarm

Depending on the interface, failure information is displayed as follows:
HART current output
Device diagnostics Device condition can be read out via HART Command 48
PROFIBUS PA
Status and alarm messages
Error current FDE (Fault Disconnection Electronic)
Diagnostics in accordance with PROFIBUS PA Profile 3.02
0 mA
PROFIBUS DP
Status and alarm messages
Diagnostics in accordance with PROFIBUS PA Profile 3.02
EtherNet/IP
Device diagnostics Device condition can be read out in Input Assembly
PROFINET
Device diagnostics According to "Application Layer protocol for decentralized periphery", Version 2.3
FOUNDATION Fieldbus
Status and alarm messages
Error current FDE (Fault Disconnection Electronic)
Diagnostics in accordance with FF-891
0 mA
Modbus RS485
Failure mode Choose from:
• NaN value instead of current value
• Last valid value
Current output 0/4 to 20 mA
4 to 20 mA
Failure mode Choose from:
• 4 to 20 mA in accordance with NAMUR recommendation NE 43
• 4 to 20 mA in accordance with US
• Min. value: 3.59 mA
• Max. value: 22.5 mA
• Freely definable value between: 3.59 to 22.5 mA
• Actual value
• Last valid value
20 Endress+Hauser
Proline Promass F 500
0 to 20 mA
Failure mode Choose from:
• Maximum alarm: 22 mA
• Freely definable value between: 0 to 20.5 mA
Pulse/frequency/switch output
Pulse output
Failure mode Choose from:
• Actual value
• No pulses
Frequency output
Failure mode Choose from:
• Actual value
• 0 Hz
• Defined value (f
Switch output
Failure mode Choose from:
• Current status
• Open
• Closed
2 to 12 500 Hz)
max
Relay output
Failure mode Choose from:
• Current status
• Open
• Closed
Local display
Plain text display With information on cause and remedial measures
Backlight Red backlighting indicates a device error.
Status signal as per NAMUR recommendation NE 107
Interface/protocol
• Via digital communication: – HART protocol – FOUNDATION Fieldbus – PROFIBUS PA – PROFIBUS DP – Modbus RS485 – EtherNet/IP – PROFINET
• Via service interface – CDI-RJ45 service interface – WLAN interface
Plain text display With information on cause and remedial measures
Additional information on remote operation →  113
Endress+Hauser 21
Web server
Plain text display With information on cause and remedial measures
Light emitting diodes (LED)
Status information Status indicated by various light emitting diodes

Ex connection data Safety-related values

Proline Promass F 500
The following information is displayed depending on the device version:
• Supply voltage active
• Data transmission active
• Device alarm/error has occurred
• EtherNet/IP network available
• EtherNet/IP connection established
• PROFINET network available
• PROFINET connection established
• PROFINET blinking feature
Order code for
Output type Safety-related values
"Output; input 1"
Option BA Current output
4 to 20 mA HART
UN = 30 V UM = 250 V
Option GA PROFIBUS PA UN = 30 V
UM = 250 V
Option LA PROFIBUS DP UN = 30 V
UM = 250 V
Option MA Modbus RS485 UN = 30 V
UM = 250 V
Option SA FOUNDATION Fieldbus UN = 30 V
UM = 250 V
Option NA EtherNet/IP UN = 30 V
UM = 250 V
Option RA PROFINET UN = 30 V
UM = 250 V
Order code for "Output; input 2"; "Output; input 3"
Output type Safety-related values
Output; input 2 Output; input 3 Output; input
"Output; input 4"
24 (+) 25 (–) 22 (+) 23 (–) 20 (+) 21 (–)
Option B Current output
4 to 20 mA
Option D User configurable input/
output
Option E Pulse/frequency/switch
output
Option F Double pulse output UN = 30 V
UN = 30 V UM = 250 V
UN = 30 V UM = 250 V
UN = 30 V UM = 250 V
DC
DC
DC
DC
UM = 250 V
Option H Relay output UN = 30 V
DC
IN =100 mADC/500 mA UM = 250 V
"Output; input 1"
26 (+) 27 (–)
DC
AC
DC
AC
DC
AC
DC
AC
DC
AC
DC
AC
DC
AC
AC
AC
AC
AC
AC
AC
1)
4
22 Endress+Hauser
Proline Promass F 500
Order code for "Output; input 2"; "Output; input 3"
Output type Safety-related values
Output; input 2 Output; input 3 Output; input
"Output; input 4"
24 (+) 25 (–) 22 (+) 23 (–) 20 (+) 21 (–)
Option I Current input
4 to 20 mA
Option J Status input UN = 30 V
UN = 30 V UM = 250 V
UM = 250 V
DC
AC
DC
AC
1) The order code "Output; input 4" is only available for the Proline 500 – digital transmitter.
Intrinsically safe values
Order code for "Output; input 1"
Option CA Current output
Option HA PROFIBUS PA Ex i Ex ia
Option TA FOUNDATION Fieldbus
Output type Intrinsically safe values
"Output; input 1"
26 (+) 27 (–)
Ui = 30 V
4 to 20 mA HART Ex i
li = 100 mA Pi = 1.25 W Li = 0 Ci = 0
Ex i
1)
Ui = 30 V li = 570 mA Pi = 8.5 W Li = 10 µH Ci = 5 nF
1)
Ex ia
Ui = 30 V li = 570 mA Pi = 8.5 W Li = 10 µH Ci = 5 nF
2)
Ex ic
Ui = 32 V li = 570 mA Pi = 8.5 W Li = 10 µH Ci = 5 nF
2)
Ex ic
Ui = 32 V li = 570 mA Pi = 8.5 W Li = 10 µH Ci = 5 nF
1)
4

Low flow cut off

Galvanic isolation

1) Only available for the Zone 1; Class I, Division 1 version
2) Only available for the Zone 2; Class I, Division 2 version and only for the Proline 500 – digital transmitter
Order code for "Output; input 2"; "Output; input 3" "Output; input 4"
Option C Current output
Output type Intrinsically safe values or NIFW values
Output; input 2 Output; input 3 Output; input 4
24 (+) 25 (–) 22 (+) 23 (–) 20 (+) 21 (–)
Ui = 30 V
4 to 20 mA Ex i
li = 100 mA Pi = 1.25 W Li = 0 Ci = 0
Option G Pulse/frequency/switch
output Ex i
Ui = 30 V li = 100 mA Pi = 1.25 W Li = 0 Ci = 0
1) The order code "Output; input 4" is only available for the Proline 500 – digital transmitter.
The switch points for low flow cut off are user-selectable.
The outputs are galvanically isolated from one another and from earth (PE).
1)
Endress+Hauser 23

Protocol-specific data HART

Manufacturer ID 0x11
Device type ID 0x3B
HART protocol revision 7
Device description files (DTM, DD)
HART load Min. 250 Ω
System integration Information on system integration: Operating Instructions →  135.
PROFIBUS PA
Manufacturer ID 0x11
Ident number 0x156D
Profile version 3.02
Device description files (GSD, DTM, DD)
Supported functions • Identification & Maintenance
Configuration of the device address
Compatibility with earlier model
System integration Information regarding system integration: Operating Instructions →  135.
Proline Promass F 500
Information and files under:
www.endress.com
• Measured variables via HART protocol
• Burst Mode functionality
Information and files under:
www.endress.com
www.profibus.org
Simplest device identification on the part of the control system and nameplate
• PROFIBUS upload/download Reading and writing parameters is up to ten times faster with PROFIBUS upload/download
• Condensed status Simplest and self-explanatory diagnostic information by categorizing diagnostic messages that occur
• DIP switches on the I/O electronics module
• Local display
• Via operating tools (e.g. FieldCare)
If the device is replaced, the measuring device Promass 500 supports the compatibility of the cyclic data with previous models. It is not necessary to adjust the engineering parameters of the PROFIBUS network with the Promass 500 GSD file.
Earlier models:
• Promass 80 PROFIBUS PA – ID No.: 1528 (hex) – Extended GSD file: EH3x1528.gsd – Standard GSD file: EH3_1528.gsd
• Promass 83 PROFIBUS PA – ID No.: 152A (hex) – Extended GSD file: EH3x152A.gsd – Standard GSD file: EH3_152A.gsd
Description of the function scope of compatibility:
Operating Instructions →  135.
• Cyclic data transmission
• Block model
• Description of the modules
PROFIBUS DP
Manufacturer ID 0x11
Ident number 0x156F
24 Endress+Hauser
Proline Promass F 500
Profile version 3.02
Device description files (GSD, DTM, DD)
Supported functions • Identification & Maintenance
Configuration of the device address
System integration Information regarding system integration: Operating Instructions →  135.
Information and files under:
www.endress.com On the product page for the device: Documents/Software → Device drivers
www.profibus.org
Simplest device identification on the part of the control system and nameplate
• PROFIBUS upload/download Reading and writing parameters is up to ten times faster with PROFIBUS upload/download
• Condensed status Simplest and self-explanatory diagnostic information by categorizing diagnostic messages that occur
• DIP switches on the I/O electronics module
• Via operating tools (e.g. FieldCare)
• Cyclic data transmission
• Block model
• Description of the modules
EtherNet/IP
Protocol • The CIP Networks Library Volume 1: Common Industrial Protocol
• The CIP Networks Library Volume 2: EtherNet/IP Adaptation of CIP
Communication type • 10Base-T
• 100Base-TX
Device profile Generic device (product type: 0x2B)
Manufacturer ID 0x11
Device type ID 0x103B
Baud rates Automatic ¹⁰⁄₁₀₀ Mbit with half-duplex and full-duplex detection
Polarity Auto-polarity for automatic correction of crossed TxD and RxD pairs
Supported CIP connections Max. 3 connections
Explicit connections Max. 6 connections
I/O connections Max. 6 connections (scanner)
Configuration options for measuring device
Configuration of the EtherNet interface
Configuration of the device address
Device Level Ring (DLR) Yes
System integration Information regarding system integration: Operating Instructions
• DIP switches on the electronics module for IP addressing
• Manufacturer-specific software (FieldCare)
• Add-on Profile Level 3 for Rockwell Automation control systems
• Web browser
• Electronic Data Sheet (EDS) integrated in the measuring device
• Speed: 10 MBit, 100 MBit, auto (factory setting)
• Duplex: half-duplex, full-duplex, auto (factory setting)
• DIP switches on the electronics module for IP addressing (last octet)
• DHCP
• Manufacturer-specific software (FieldCare)
• Add-on Profile Level 3 for Rockwell Automation control systems
• Web browser
• EtherNet/IP tools, e.g. RSLinx (Rockwell Automation)
→  135.
• Cyclic data transmission
• Block model
• Input and output groups
Endress+Hauser 25
Proline Promass F 500
PROFINET
Protocol "Application layer protocol for decentral device periphery and distributed
automation", version 2.3
Communication type 100 MBit/s
Conformity class Conformance Class B
Netload Class Netload Class II
Baud rates Automatic 100 Mbit/s with full-duplex detection
Cycle times From 8 ms
Polarity Auto-polarity for automatic correction of crossed TxD and RxD pairs
Media Redundancy Protocol (MRP)
Device profile Application interface identifier 0xF600
Manufacturer ID 0x11
Device type ID 0x843B
Device description files (GSD, DTM, DD)
Supported connections • 1 x AR (IO Controller AR)
Configuration options for measuring device
Configuration of the device name
Supported functions • Identification & Maintenance
System integration Information regarding system integration: Operating Instructions →  135.
Yes
Generic device
Information and files under:
www.endress.com On the product page for the device: Documents/Software → Device drivers
www.profibus.org
• 1 x AR (IO-Supervisor Device AR connection allowed)
• 1 x Input CR (Communication Relation)
• 1 x Output CR (Communication Relation)
• 1 x Alarm CR (Communication Relation)
• DIP switches on the electronics module, for device name assignment (last part)
• Manufacturer-specific software (FieldCare, DeviceCare)
• Web browser
• Device master file (GSD), can be read out via the integrated Web server of the measuring device
• DIP switches on the electronics module, for device name assignment (last part)
• DCP protocol
• Process Device Manager (PDM)
• Integrated Web server
Simple device identification via: – Control system – Nameplate
• Measured value status The process variables are communicated with a measured value status
• Blinking feature via the onsite display for simple device identification and assignment
• Device operation via operating tools (e.g. FieldCare, DeviceCare, SIMATIC PDM)
• Cyclic data transmission
• Overview and description of the modules
• Status coding
• Startup configuration
• Factory setting:
FOUNDATION Fieldbus
Manufacturer ID 0x452B48 (hex)
Ident number 0x103B (hex)
26 Endress+Hauser
Proline Promass F 500
Device revision 1
DD revision Information and files under:
CFF revision
Interoperability Test Kit (ITK) Version 6.2.0
ITK Test Campaign Number Information:
Link Master capability (LAS) Yes
Choice of "Link Master" and "Basic Device"
Node address Factory setting: 247 (0xF7)
Supported functions The following methods are supported:
Virtual Communication Relationships (VCRs)
Number of VCRs 44
Number of link objects in VFD 50
Permanent entries 1
Client VCRs 0
Server VCRs 10
Source VCRs 43
Sink VCRs 0
Subscriber VCRs 43
Publisher VCRs 43
Device Link Capabilities
Slot time 4
Min. delay between PDU 8
Max. response delay 16
System integration Information regarding system integration: Operating Instructions →  135.
www.endress.com
www.fieldbus.org
www.endress.com
www.fieldbus.org
Yes Factory setting: Basic Device
• Restart
• ENP Restart
• Diagnostic
• Set to OOS
• Set to AUTO
• Read trend data
• Read event logbook
• Cyclic data transmission
• Description of the modules
• Execution times
• Methods
Modbus RS485
Protocol Modbus Applications Protocol Specification V1.1
Response times • Direct data access: typically 25 to 50 ms
• Auto-scan buffer (data range): typically 3 to 5 ms
Device type Slave
Slave address range 1 to 247
Broadcast address range 0
Endress+Hauser 27
Proline Promass F 500
Function codes • 03: Read holding register
• 04: Read input register
• 06: Write single registers
• 08: Diagnostics
• 16: Write multiple registers
• 23: Read/write multiple registers
Broadcast messages Supported by the following function codes:
• 06: Write single registers
• 16: Write multiple registers
• 23: Read/write multiple registers
Supported baud rate • 1 200 BAUD
• 2 400 BAUD
• 4 800 BAUD
• 9 600 BAUD
• 19 200 BAUD
• 38 400 BAUD
• 57 600 BAUD
• 115 200 BAUD
Data transfer mode • ASCII
• RTU
Data access Each device parameter can be accessed via Modbus RS485.
For Modbus register information
Compatibility with earlier model
System integration Information on system integration: Operating Instructions →  135.
If the device is replaced, the measuring device Promass 500 supports the compatibility of the Modbus registers for the process variables and the diagnostic information with the previous model Promass 83. It is not necessary to change the engineering parameters in the automation system.
Description of the function scope of compatibility:
Operating Instructions →  135.
• Modbus RS485 information
• Function codes
• Register information
• Response time
• Modbus data map

Power supply

Terminal assignment Transmitter: supply voltage, input/outputs

HART
Supply voltage Input/output
1
1 (+) 2 (–) 26 (+) 27 (–) 24 (+) 25 (–) 22 (+) 23 (–) 20 (+) 21 (–)
The terminal assignment depends on the specific device version ordered →  15.
Input/output
Input/output
2
3
Input/output
4
FOUNDATION Fieldbus
Supply voltage Input/output
1
1 (+) 2 (–) 26 (A) 27 (B) 24 (+) 25 (–) 22 (+) 23 (–) 20 (+) 21 (–)
The terminal assignment depends on the specific device version ordered →  15.
Input/output
2
Input/output
3
Input/output
4
28 Endress+Hauser
Proline Promass F 500
PROFIBUS PA
Supply voltage Input/output
1
1 (+) 2 (–) 26 (B) 27 (A) 24 (+) 25 (–) 22 (+) 23 (–) 20 (+) 21 (–)
The terminal assignment depends on the specific device version ordered →  15.
Input/output
2
Input/output
3
Input/output
4
PROFIBUS DP
Supply voltage Input/output
1
1 (+) 2 (–) 26 (B) 27 (A) 24 (+) 25 (–) 22 (+) 23 (–) 20 (+) 21 (–)
The terminal assignment depends on the specific device version ordered →  15.
Input/output
2
Input/output
3
Input/output
4
Modbus RS485
Supply voltage Input/output
1
1 (+) 2 (–) 26 (B) 27 (A) 24 (+) 25 (–) 22 (+) 23 (–) 20 (+) 21 (–)
The terminal assignment depends on the specific device version ordered →  15.
Input/output
2
Input/output
3
Input/output
4
PROFINET
Supply voltage Input/output
1
1 (+) 2 (–) PROFINET
(RJ45 connector)
Input/output
2
24 (+) 25 (–) 22 (+) 23 (–) 20 (+) 21 (–)
The terminal assignment depends on the specific device version
Input/output
3
ordered →  15.
Input/output
4

Device plugs available

EtherNet/IP
Supply voltage Input/output
1
1 (+) 2 (–) EtherNet/IP
(RJ45 connector)
Input/output
2
24 (+) 25 (–) 22 (+) 23 (–) 20 (+) 21 (–)
The terminal assignment depends on the specific device version
Input/output
3
ordered →  15.
Input/output
4
Transmitter and sensor connection housing: connecting cable
The sensor and transmitter, which are mounted in separate locations, are interconnected by a connecting cable. The cable is connected via the sensor connection housing and the transmitter housing.
Terminal assignment and connection of the connecting cable:
• Proline 500 – digital→  33
• Proline 500 →  34
Device plugs may not be used in hazardous areas!
Device plugs for fieldbus systems:
Order code for "Input; output 1"
• Option SA "FOUNDATION Fieldbus" →  30
• Option GA "PROFIBUS PA" →  30
• Option RA "PROFINET" →  30
• Option NA "EtherNet/IP" →  30
Device plug for connecting to the service interface:
Order code for "Accessory mounted" option NB, adapter RJ45 M12 (service interface) →  31
Endress+Hauser 29
Order code for "Input; output 1", option SA "FOUNDATION Fieldbus"
1
2
4
3
Order code for Cable entry/connection →  34
"Electrical connection" 2 3
M, 3, 4, 5 7/8" connector
Order code for "Input; output 1", option GA "PROFIBUS PA"
Order code for Cable entry/connection →  34
"Electrical connection" 2 3
L, N, P, U Connector M12 × 1
Order code for "Input; output 1", option RA "PROFINET"
Order code for Cable entry/connection →  34
"Electrical connection" 2 3
L, N, P, U Connector M12 × 1
1) 2)
1) 2)
1) 2)
R
, S
, T
, V
1) 2)
Connector M12 × 1 Connector M12 × 1
Proline Promass F 500
1) Cannot be combined with an external WLAN antenna (order code for "Enclosed accessories", option P8) of an RJ45 M12 adapter for the service interface (order code for "Accessories mounted", option NB) or of the remote display and operating module DKX001.
2) Suitable for integrating the device in a ring topology.
Order code for "Input; output 1", option NA "EtherNet/IP"
Order code for Cable entry/connection →  34
"Electrical connection" 2 3
L, N, P, U Connector M12 × 1
1) 2)
1) 2)
1) 2)
R
, S
, T
, V
1) 2)
Connector M12 × 1 Connector M12 × 1
1) Cannot be combined with an external WLAN antenna (order code for "Enclosed accessories", option P8) of an RJ45 M12 adapter for the service interface (order code for "Accessories mounted", option NB) or of the remote display and operating module DKX001
2) Suitable for integrating the device in a ring topology.
Order code for "Accessory mounted", option NB "Adapter RJ45 M12 (service interface)"
Order code Cable entry/coupling →  34
"Accessory mounted" Cable entry
2
NB Plug M12 × 1
Cable entry
3

Pin assignment, device plug FOUNDATION Fieldbus

Pin Assignment Coding Plug/socket
1 + Signal + A Plug
2 - Signal –
3 Grounding
4 Not assigned
30 Endress+Hauser