A grounded terminal which, as far as the operator is concerned, is grounded via a
grounding system.
Protective Earth (PE)
A terminal which must be connected to ground prior to establishing any other
connections.
The ground terminals are situated inside and outside the device:
• Inner ground terminal: Connects the protectiv earth to the mains supply.
• Outer ground terminal: Connects the device to the plant grounding system.
Communication symbols
SymbolMeaning
Wireless Local Area Network (WLAN)
Communication via a wireless, local network.
LED
Light emitting diode is off.
LED
Light emitting diode is on.
LED
Light emitting diode is flashing.
Symbols for certain types of information
SymbolMeaning
Permitted
Procedures, processes or actions that are permitted.
Preferred
Procedures, processes or actions that are preferred.
Forbidden
Procedures, processes or actions that are forbidden.
Tip
Indicates additional information.
Reference to documentation.
Reference to page.
Reference to graphic.
Visual inspection.
4Endress+Hauser
Page 5
Proline Promass F 300
1.
-
.
213
Symbols in graphics
SymbolMeaning
1, 2, 3, ...Item numbers
, 2., 3., …Series of steps
A, B, C, ...Views
A-A, B-B, C-C, ...Sections
Hazardous area
Safe area (non-hazardous area)
Flow direction
Function and system design
Measuring principle
The measuring principle is based on the controlled generation of Coriolis forces. These forces are
always present in a system when both translational and rotational movements are superimposed.
Fc = 2 · ∆m (ν · ω)
Fc = Coriolis force
∆m = moving mass
ω = rotational velocity
ν = radial velocity in rotating or oscillating system
The amplitude of the Coriolis force depends on the moving mass ∆m, its velocity ν in the system and
thus on the mass flow. Instead of a constant rotational velocity ω, the sensor uses oscillation.
In the sensor, two parallel measuring tubes containing flowing fluid oscillate in antiphase, acting like
a tuning fork. The Coriolis forces produced at the measuring tubes cause a phase shift in the tube
oscillations (see illustration):
• At zero flow (when the fluid is at a standstill) the two tubes oscillate in phase (1).
• Mass flow causes deceleration of the oscillation at the inlet of the tubes (2) and acceleration at the
outlet (3).
A0028850
The phase difference (A-B) increases with increasing mass flow. Electrodynamic sensors register the
tube oscillations at the inlet and outlet. System balance is ensured by the antiphase oscillation of the
two measuring tubes. The measuring principle operates independently of temperature, pressure,
viscosity, conductivity and flow profile.
Endress+Hauser5
Page 6
Proline Promass F 300
Density measurement
The measuring tube is continuously excited at its resonance frequency. A change in the mass and
thus the density of the oscillating system (comprising measuring tube and fluid) results in a
corresponding, automatic adjustment in the oscillation frequency. Resonance frequency is thus a
function of medium density. The microprocessor utilizes this relationship to obtain a density signal.
Volume measurement
Together with the measured mass flow, this is used to calculate the volume flow.
Temperature measurement
The temperature of the measuring tube is determined in order to calculate the compensation factor
due to temperature effects. This signal corresponds to the process temperature and is also available
as an output signal.
6Endress+Hauser
Page 7
Proline Promass F 300
Measuring system
The device consists of a transmitter and a sensor.
The device is available as a compact version:
The transmitter and sensor form a mechanical unit.
• External operation via 4-line, backlit, graphic local display with touch
control and guided menus ("Make-it-run" wizards) for applicationspecific commissioning.
• Via service interface or WLAN interface:
– Operating tools (e.g. FieldCare, DeviceCare, SmartBlue app)
– Web server (access via Web browser, e.g. Microsoft Internet
Explorer, Microsoft Edge)
Sensor
PromassF• Excellent performance across a wide range of applications
• Simultaneous measurement of flow, volume flow, density and
temperature (multivariable)
• Immune to process influences
• Nominal diameter range: DN 8 to 250 (³⁄₈ to 10")
– Process connections: stainless steel, 1.4404 (316/316L); 1.4301
(304); Alloy C22, 2.4602 (UNS N06022)
Endress+Hauser7
Page 8
Equipment architecture
2
1
6
5
7
4
3
3
Proline Promass F 300
A0027512
1Possibilities for integrating measuring devices into a system
1Control system (e.g. PLC)
2Connecting cable (0/4 to 20 mA HART etc.)
3Fieldbus
4Segment coupler
5Non-hazardous area
6Hazardous area: Zone 2; Class I, Division 2
7Hazardous area: Zone 1; Class I, Division 1
SafetyIT security
Our warranty is valid only if the device is installed and used as described in the Operating
Instructions. The device is equipped with security mechanisms to protect it against any inadvertent
changes to the settings.
IT security measures, which provide additional protection for the device and associated data transfer,
must be implemented by the operators themselves in line with their security standards.
Device-specific IT security
The device offers a range of specific functions to support protective measures on the operator's side.
These functions can be configured by the user and guarantee greater in-operation safety if used
correctly. An overview of the most important functions is provided in the following section.
Function/interfaceFactory settingRecommendation
Write protection via hardware write
protection switch → 9
Access code
(also applies for Web server login or
FieldCare connection) → 9
WLAN
(order option in display module)
WLAN security modeEnabled (WPA2-
WLAN passphrase
(password) → 9
WLAN modeAccess PointOn an individual basis following risk
Web server→ 9Enabled.On an individual basis following risk
CDI-RJ45 service interface → 10–On an individual basis following risk
Not enabled.On an individual basis following risk
assessment.
Not enabled
(0000).
Enabled.On an individual basis following risk
PSK)
Serial numberAssign a customized access code during
Assign a customized access code during
commissioning.
assessment.
Do not change.
commissioning.
assessment.
assessment.
assessment.
8Endress+Hauser
Page 9
Proline Promass F 300
Protecting access via hardware write protection
Write access to the device parameters via the local display, Web browser or operating tool (e.g.
FieldCare, DeviceCare) can be disabled via a write protection switch (DIP switch on the
motherboard). When hardware write protection is enabled, only read access to the parameters is
possible.
Hardware write protection is disabled when the device is delivered.
Protecting access via a password
Different passwords are available to protect write access to the device parameters or access to the
device via the WLAN interface.
• User-specific access code
Protect write access to the device parameters via the local display, Web browser or operating tool
(e.g. FieldCare, DeviceCare). Access authorization is clearly regulated through the use of a userspecific access code.
• WLAN passphrase
The network key protects a connection between an operating unit (e.g. notebook or tablet) and the
device via the WLAN interface which can be ordered as an option.
• Infrastructure mode
When the device is operated in infrastructure mode, the WLAN passphrase corresponds to the
WLAN passphrase configured on the operator side.
User-specific access code
Write access to the device parameters via the local display, Web browser or operating tool (e.g.
FieldCare, DeviceCare) can be protected by the modifiable, user-specific access code.
WLAN passphrase: Operation as WLAN access point
A connection between an operating unit (e.g. notebook or tablet) and the device via the WLAN
interface, which can be ordered as an optional extra, is protected by the network key. The WLAN
authentication of the network key complies with the IEEE 802.11 standard.
When the device is delivered, the network key is pre-defined depending on the device. It can be
changed via the WLAN settings submenu in the WLAN passphrase parameter.
Infrastructure mode
A connection between the device and WLAN access point is protected by means of an SSID and
passphrase on the system side. Please contact the relevant system administrator for access.
General notes on the use of passwords
• The access code and network key supplied with the device should be changed during
commissioning.
• Follow the general rules for generating a secure password when defining and managing the access
code or network key.
• The user is responsible for the management and careful handling of the access code and network
key.
Access via Web server
The device can be operated and configured via a Web browser with the integrated Web server. The
connection is via the service interface (CDI-RJ45) or the WLAN interface. For device versions with
the EtherNet/IP and PROFINET communication protocols, the connection can also be established via
the terminal connection for signal transmission with EtherNet/IP or PROFINET (RJ45 connector).
The Web server is enabled when the device is delivered. The Web server can be disabled if necessary
(e.g. after commissioning) via the Web server functionality parameter.
The device and status information can be hidden on the login page. This prevents unauthorized
access to the information.
For detailed information on device parameters, see:
The "Description of Device Parameters" document → 120
Access via OPC-UA
The "OPC UA Server" application package is available in the device version with the HART
communication protocol → 117.
Endress+Hauser9
Page 10
Proline Promass F 300
The device can communicate with OPC UA clients using the "OPC UA Server" application package.
The OPC UA server integrated in the device can be accessed via the WLAN access point using the
WLAN interface - which can be ordered as an optional extra - or the service interface (CDI- RJ45) via
Ethernet network. Access rights and authorization as per separate configuration.
The following Security Modes are supported as per the OPC UA Specification (IEC 62541):
• None
• Basic128Rsa15 – signed
• Basic128Rsa15 – signed and encrypted
Access via service interface (CDI-RJ45)
The device can be connected to a network via the service interface (CDI-RJ45). Device-specific
functions guarantee the secure operation of the device in a network.
The use of relevant industrial standards and guidelines that have been defined by national and
international safety committees, such as IEC/ISA62443 or the IEEE, is recommended. This includes
organizational security measures such as the assignment of access authorization as well as technical
measures such as network segmentation.
The device can be integrated in a ring topology. The device is integrated via the terminal
connection for signal transmission (output 1) and the connection to the service interface (CDIRJ45) → 105.
10Endress+Hauser
Page 11
Proline Promass F 300
Input
Measured variableDirect measured variables
• Mass flow
• Density
• Temperature
Calculated measured variables
• Volume flow
• Corrected volume flow
• Reference density
Measuring rangeMeasuring ranges for liquids
DNMeasuring range full scale values
min(F)
to
[mm][in][kg/h][lb/min]
8³⁄₈0 to 20000 to 73.50
15½0 to 65000 to 238.9
2510 to 180000 to 661.5
401½0 to 450000 to 1654
5020 to 700000 to 2573
8030 to 1800000 to 6615
10040 to 3500000 to 12860
15060 to 8000000 to 29400
250100 to 22000000 to 80850
Measuring range for gases
The full scale value depends on the density and the sound velocity of the gas used and can be
calculated with the formula below:
= minimum (
max(G)
max(G)
max(F)
<
ρ
max(G)
G
max(F)
xConstant dependent on nominal diameter
c
G
d
i
· ρG : x ; ρG · cG · π/2 · (di)2 · 3600)
max(F)
Maximum full scale value for gas [kg/h]
Maximum full scale value for liquid [kg/h]
can never be greater than
max(G)
Gas density in [kg/m³] at operating conditions
Sound velocity (gas) [m/s]
Measuring tube internal diameter [m]
max(F)
max(F)
DNx
[mm][in][kg/m3]
8³⁄₈60
15½80
25190
401½90
50290
803110
Endress+Hauser11
Page 12
DNx
[mm][in][kg/m3]
1004130
1506200
25010200
To calculate the measuring range, use the Applicator sizing tool → 119
Calculation example for gas
• Sensor: Promass F, DN 50
• Gas: Air with a density of 60.3 kg/m³ (at 20 °C and 50 bar)
• Measuring range (liquid): 70000 kg/h
• x = 90 kg/m³ (for Promass F, DN 50)
Maximum possible full scale value:
Flow rates above the preset full scale value do not override the electronics unit, with the result that
the totalizer values are registered correctly.
Input signalInput and output versions
→ 14
External measured values
To increase the accuracy of certain measured variables or to calculate the corrected volume flow for
gases, the automation system can continuously write different measured values to the measuring
device:
• Operating pressure to increase accuracy (Endress+Hauser recommends the use of a pressure
measuring device for absolute pressure, e.g. Cerabar M or Cerabar S)
• Medium temperature to increase accuracy (e.g. iTEMP)
• Reference density for calculating the corrected volume flow for gases
Various pressure transmitters and temperature measuring devices can be ordered from Endress
+Hauser: see "Accessories" section → 119
It is recommended to read in external measured values to calculate the corrected volume flow.
HART protocol
The measured values are written from the automation system to the measuring device via the HART
protocol. The pressure transmitter must support the following protocol-specific functions:
• HART protocol
• Burst mode
Current input
The measured values are written from the automation system to the measuring device via the
current input → 13.
Digital communication
The measured values can be written from the automation system to the measuring via:
• FOUNDATION Fieldbus
• PROFIBUS DP
• PROFIBUS PA
• Modbus RS485
• EtherNet/IP
• PROFINET
12Endress+Hauser
Page 13
Proline Promass F 300
Current input 0/4 to 20 mA
Current input0/4 to 20 mA (active/passive)
Current span• 4 to 20 mA (active)
• 0/4 to 20 mA (passive)
Resolution1 µA
Voltage dropTypically: 0.6 to 2 V for 3.6 to 22 mA (passive)
Maximum input voltage≤ 30 V (passive)
Open-circuit voltage≤ 28.8 V (active)
Possible input variables• Pressure
• Temperature
• Density
Status input
Maximum input values• DC –3 to 30 V
• If status input is active (ON): Ri >3 kΩ
Response timeAdjustable: 5 to 200 ms
Input signal level• Low signal: DC –3 to +5 V
• High signal: DC 12 to 30 V
Assignable functions• Off
• Reset the individual totalizers separately
• Reset all totalizers
• Flow override
Endress+Hauser13
Page 14
Output
Proline Promass F 300
Output and input variants
Depending on the option selected for output/input 1, different options are available for the other
outputs and inputs. Only one option can be selected for each output/input 1 to 3. The table must be
read vertically (↓).
Example: If the option BA "4–20 mA HART" was selected for output/input 1, one of the options A, B,
D, E, F, H, I or J is available for output 2 and one of the options A, B, D, E, F, H, I or J is available for
output 3.
Order code for "Output; input 1" (020) →Possible options
Current output 4 to 20 mA HARTBA
Current output 4 to 20 mA HART Ex i↓CA
FOUNDATION Fieldbus↓SA
FOUNDATION Fieldbus Ex i↓TA
PROFIBUS DP↓LA
PROFIBUS PA↓GA
PROFIBUS PA Ex i↓HA
Modbus RS485↓MA
EtherNet/IP 2-port switch integrated↓NA
PROFINET 2-port switch integrated↓RA
Order code for "Output; input 2" (021) →↓↓↓↓↓↓↓↓↓↓
Not assignedAAAAAAAAAA
Current output 0/4 to 20 mABBBBBBB
Current output 0/4 to 20 mA (Ex i)CCC
User configurable input/output
Pulse/frequency/switch outputEEEEEEE
Double pulse output
Pulse/frequency/switch output (Ex i)GGG
Relay outputHHHHHHH
Current input 0/4 to 20 mAIIIIIII
Status inputJJJJJJJ
Order code for "Output; input 3" (022) →↓↓↓↓↓↓↓↓↓↓
Not assignedAAAAAAAAAA
Current output 0/4 to 20 mABBBBB
Current output 0/4 to 20 mA (Ex i)C
User configurable input/outputDDDDD
Pulse/frequency/switch outputEEEEE
Double pulse output (slave)FF
Pulse/frequency/switch output (Ex i)G
Relay outputHHHHH
Current input 0/4 to 20 mAIIIII
Status inputJJJJJ
2)
1)
DDDDDDD
FF
1)A specific input or output can be assigned to a user configurable input/output → 18.
2)If double pulse output (F) is selected for output/input 2 (021), only the double pulse output (F) option is available for selection for output/input 3
(022).
14Endress+Hauser
Page 15
Proline Promass F 300
Output signalHART current output
Current output4 to 20 mA HART
Current spanCan be set to: 4 to 20 mA (active/passive)
Open-circuit voltageDC 28.8 V (active)
Maximum input voltageDC 30 V (passive)
Load250 to 700 Ω
Resolution0.38 µA
DampingConfigurable: 0.07 to 999 s
Assignable measured
variables
Ex-i, passive
• Mass flow
• Volume flow
• Corrected volume flow
• Density
• Reference density
• Temperature
• Electronic temperature
• Oscillation frequency 0
• Oscillation damping 0
• Signal asymmetry
• Exciter current 0
The range of options increases if the measuring device has one or more
application packages.
PROFIBUS PA
PROFIBUS PAIn accordance with EN 50170 Volume 2, IEC 61158-2 (MBP), galvanically
isolated
Data transmission31.25 kbit/s
Current consumption10 mA
16 mA
Permitted supply voltage9 to 32 V
Bus connectionWith integrated reverse polarity protection
PROFIBUS DP
Signal encodingNRZ code
Data transfer9.6 kBaud…12 MBaud
EtherNet/IP
StandardsIn accordance with IEEE 802.3
PROFINET
StandardsIn accordance with IEEE 802.3
FOUNDATION Fieldbus
FOUNDATION FieldbusH1, IEC 61158-2, galvanically isolated
Data transfer31.25 kbit/s
Endress+Hauser15
Page 16
Current consumption10 mA
Permitted supply voltage9 to 32 V
Bus connectionWith integrated reverse polarity protection
Modbus RS485
Physical interfaceRS485 in accordance with EIA/TIA-485 standard
Terminating resistorIntegrated, can be activated via DIP switches
Current output 0/4 to 20 mA
Current output0/4 to 20 mA
Maximum output values22.5 mA
Current spanCan be set to:
• 4 to 20 mA (active)
• 0/4 to 20 mA (passive)
Ex-i, passive
Proline Promass F 300
Open-circuit voltageDC 28.8 V (active)
Maximum input voltageDC 30 V (passive)
Load0 to 700 Ω
Resolution0.38 µA
DampingAdjustable: 0.07 to 999 s
Assignable measured
variables
• Mass flow
• Volume flow
• Corrected volume flow
• Density
• Reference density
• Temperature
• Electronic temperature
• Oscillation frequency 0
• Oscillation damping 0
• Signal asymmetry
• Exciter current 0
The range of options increases if the measuring device has one or more
application packages.
Pulse/frequency/switch output
FunctionCan be set to pulse, frequency or switch output
VersionOpen collector
Can be set to:
• Active
• Passive
Ex-i, passive
Maximum input valuesDC 30 V, 250 mA (passive)
Open-circuit voltageDC 28.8 V (active)
Voltage dropFor 22.5 mA: ≤ DC 2 V
Pulse output
Maximum input valuesDC 30 V, 250 mA (passive)
Maximum output current22.5 mA (active)
16Endress+Hauser
Page 17
Proline Promass F 300
Open-circuit voltageDC 28.8 V (active)
Pulse widthAdjustable: 0.05 to 2000 ms
Maximum pulse rate10000 Impulse/s
Pulse valueAdjustable
Assignable measured
variables
Frequency output
Maximum input valuesDC 30 V, 250 mA (passive)
Maximum output current22.5 mA (active)
Open-circuit voltageDC 28.8 V (active)
Output frequencyAdjustable: end value frequency 2 to 10000 Hz (f
DampingAdjustable: 0 to 999 s
Pulse/pause ratio1:1
Assignable measured
variables
Switch output
Maximum input valuesDC 30 V, 250 mA (passive)
Open-circuit voltageDC 28.8 V (active)
Switching behaviorBinary, conductive or non-conductive
Switching delayAdjustable: 0 to 100 s
Number of switching
cycles
Assignable functions• Off
• Mass flow
• Volume flow
• Corrected volume flow
• Mass flow
• Volume flow
• Corrected volume flow
• Density
• Reference density
• Temperature
• Electronic temperature
• Oscillation frequency 0
• Oscillation damping 0
• Signal asymmetry
• Exciter current 0
The range of options increases if the measuring device has one or more
application packages.
Unlimited
• On
• Diagnostic behavior
• Limit value
– Mass flow
– Volume flow
– Corrected volume flow
– Density
– Reference density
– Temperature
– Totalizer 1-3
• Flow direction monitoring
• Status
– Partially filled pipe detection
– Low flow
The range of options increases if the measuring device has one or more
application packages.
= 12500 Hz)
max
Endress+Hauser17
Page 18
Double pulse output
FunctionDouble pulse
VersionOpen collector
Can be set to:
• Active
• Passive
Maximum input valuesDC 30 V, 250 mA (passive)
Open-circuit voltageDC 28.8 V (active)
Voltage dropFor 22.5 mA: ≤ DC 2 V
Output frequencyAdjustable: 0 to 1000 Hz
DampingAdjustable: 0 to 999 s
Pulse/pause ratio1:1
Assignable measured
variables
• Mass flow
• Volume flow
• Corrected volume flow
• Density
• Reference density
• Temperature
The range of options increases if the measuring device has one or more
application packages.
Proline Promass F 300
Relay output
FunctionSwitch output
VersionRelay output, galvanically isolated
Switching behaviorCan be set to:
• NO (normally open), factory setting
• NC (normally closed)
Maximum switching
capacity (passive)
Assignable functions• Off
• DC 30 V, 0.1 A
• AC 30 V, 0.5 A
• On
• Diagnostic behavior
• Limit value
– Mass flow
– Volume flow
– Corrected volume flow
– Density
– Reference density
– Temperature
– Totalizer 1-3
• Flow direction monitoring
• Status
– Partially filled pipe detection
– Low flow
The range of options increases if the measuring device has one or more
application packages.
User configurable input/output
One specific input or output is assigned to a user-configurable input/output (configurable I/O)
during device commissioning.
The following inputs and outputs are available for assignment:
• Choice of current output: 4 to 20 mA (active), 0/4 to 20 mA (passive)
• Pulse/frequency/switch output
• Choice of current input: 4 to 20 mA (active), 0/4 to 20 mA (passive)
• Status input
18Endress+Hauser
Page 19
Proline Promass F 300
The technical values correspond to those of the inputs and outputs described in this section.
Signal on alarm
Depending on the interface, failure information is displayed as follows:
HART current output
Device diagnosticsDevice condition can be read out via HART Command 48
PROFIBUS PA
Status and alarm
messages
Error current FDE (Fault
Disconnection Electronic)
Diagnostics in accordance with PROFIBUS PA Profile 3.02
0 mA
PROFIBUS DP
Status and alarm
messages
Diagnostics in accordance with PROFIBUS PA Profile 3.02
EtherNet/IP
Device diagnosticsDevice condition can be read out in Input Assembly
PROFINET
Device diagnosticsAccording to "Application Layer protocol for decentralized periphery", Version 2.3
FOUNDATION Fieldbus
Status and alarm
messages
Error current FDE (Fault
Disconnection Electronic)
Diagnostics in accordance with FF-891
0 mA
Modbus RS485
Failure modeChoose from:
• NaN value instead of current value
• Last valid value
Current output 0/4 to 20 mA
4 to 20 mA
Failure modeChoose from:
• 4 to 20 mA in accordance with NAMUR recommendation NE 43
• 4 to 20 mA in accordance with US
• Min. value: 3.59 mA
• Max. value: 22.5 mA
• Freely definable value between: 3.59 to 22.5 mA
• Actual value
• Last valid value
Endress+Hauser19
Page 20
0 to 20 mA
Failure modeChoose from:
• Maximum alarm: 22 mA
• Freely definable value between: 0 to 20.5 mA
Pulse/frequency/switch output
Pulse output
Failure modeChoose from:
• Actual value
• No pulses
Frequency output
Failure modeChoose from:
• Actual value
• 0 Hz
• Defined value (f
Switch output
Failure modeChoose from:
• Current status
• Open
• Closed
2 to 12500 Hz)
max
Proline Promass F 300
Relay output
Failure modeChoose from:
• Current status
• Open
• Closed
Local display
Plain text displayWith information on cause and remedial measures
BacklightRed backlighting indicates a device error.
Status signal as per NAMUR recommendation NE 107
Interface/protocol
• Via digital communication:
– HART protocol
– FOUNDATION Fieldbus
– PROFIBUS PA
– PROFIBUS DP
– Modbus RS485
– EtherNet/IP
– PROFINET
• Via service interface
– CDI-RJ45 service interface
– WLAN interface
Plain text displayWith information on cause and remedial measures
Additional information on remote operation → 101
20Endress+Hauser
Page 21
Proline Promass F 300
Web server
Plain text displayWith information on cause and remedial measures
Light emitting diodes (LED)
Status informationStatus indicated by various light emitting diodes
Ex connection dataSafety-related values
The following information is displayed depending on the device version:
• Supply voltage active
• Data transmission active
• Device alarm/error has occurred
• EtherNet/IP network available
• EtherNet/IP connection established
• PROFINET network available
• PROFINET connection established
• PROFINET blinking feature
Order code for
Output typeSafety-related values
"Output; input 1"
Option BACurrent output
4 to 20 mA HART
UN = 30 V
UM = 250 V
Option GAPROFIBUS PAUN = 30 V
UM = 250 V
Option LAPROFIBUS DPUN = 30 V
UM = 250 V
Option MAModbus RS485UN = 30 V
UM = 250 V
Option SAFOUNDATION FieldbusUN = 30 V
UM = 250 V
Option NAEtherNet/IPUN = 30 V
UM = 250 V
Option RAPROFINETUN = 30 V
UM = 250 V
Order code for
"Output; input 2";
"Output; input 3"
Output typeSafety-related values
Output; input 2Output; input 3
24 (+)25 (–)22 (+)23 (–)
Option BCurrent output
4 to 20 mA
Option DUser configurable input/
output
Option EPulse/frequency/switch
output
UN = 30 V
UM = 250 V
UN = 30 V
UM = 250 V
UN = 30 V
UM = 250 V
Option FDouble pulse outputUN = 30 V
UM = 250 V
Option HRelay outputUN = 30 V
IN =100 mADC/500 mA
UM = 250 V
Option ICurrent input 4 to 20 mA UN = 30 V
UM = 250 V
Option JStatus inputUN = 30 V
UM = 250 V
"Output; input 1"
26 (+)27 (–)
DC
AC
DC
AC
DC
AC
DC
AC
DC
AC
DC
AC
DC
AC
DC
AC
DC
AC
DC
AC
DC
AC
DC
AC
AC
DC
AC
DC
AC
Endress+Hauser21
Page 22
Intrinsically safe values
Proline Promass F 300
Order code for
"Output; input 1"
Option CACurrent output
Option HAPROFIBUS PA Ex iEx ia
Option TAFOUNDATION Fieldbus
1)Only available for the Zone 1; Class I, Division 1 version
2)Only available for the Zone 2; Class I, Division 2 version transmitter
Order code for
"Output; input 2";
"Output; input 3"
Option CCurrent output
Option GPulse/frequency/switch
Output typeIntrinsically safe values
26 (+)27 (–)
Ui = 30 V
4 to 20 mA HART Ex i
Ex i
Output typeIntrinsically safe values or NIFW values
4 to 20 mA Ex i
output Ex i
li = 100 mA
Pi = 1.25 W
Li = 0
Ci = 0
1)
Ui = 30 V
li = 570 mA
Pi = 8.5 W
Li = 10 µH
Ci = 5 nF
1)
Ex ia
Ui = 30 V
li = 570 mA
Pi = 8.5 W
Li = 10 µH
Ci = 5 nF
Output; input 2Output; input 3
24 (+)25 (–)22 (+)23 (–)
Ui = 30 V
li = 100 mA
Pi = 1.25 W
Li = 0
Ci = 0
Ui = 30 V
li = 100 mA
Pi = 1.25 W
Li = 0
Ci = 0
"Output; input 1"
2)
Ex ic
Ui = 32 V
li = 570 mA
Pi = 8.5 W
Li = 10 µH
Ci = 5 nF
2)
Ex ic
Ui = 32 V
li = 570 mA
Pi = 8.5 W
Li = 10 µH
Ci = 5 nF
Low flow cut off
Galvanic isolation
The switch points for low flow cut off are user-selectable.
The outputs are galvanically isolated from one another and from earth (PE).
Protocol-specific dataHART
Manufacturer ID0x11
Device type ID0x3B
HART protocol revision7
Device description files
(DTM, DD)
HART loadMin. 250 Ω
System integrationInformation on system integration: Operating Instructions → 120.
Information and files under:
www.endress.com
• Measured variables via HART protocol
• Burst Mode functionality
22Endress+Hauser
Page 23
Proline Promass F 300
PROFIBUS PA
Manufacturer ID0x11
Ident number0x156D
Profile version3.02
Device description files (GSD,
DTM, DD)
Supported functions• Identification & Maintenance
Configuration of the device
address
Compatibility with
earlier model
System integrationInformation regarding system integration: Operating Instructions → 120.
Information and files under:
• www.endress.com
• www.profibus.org
Simplest device identification on the part of the control system and
nameplate
• PROFIBUS upload/download
Reading and writing parameters is up to ten times faster with PROFIBUS
upload/download
• Condensed status
Simplest and self-explanatory diagnostic information by categorizing
diagnostic messages that occur
• DIP switches on the I/O electronics module
• Local display
• Via operating tools (e.g. FieldCare)
If the device is replaced, the measuring device Promass 300 supports the
compatibility of the cyclic data with previous models. It is not necessary to
adjust the engineering parameters of the PROFIBUS network with the
Promass 300 GSD file.
Earlier models:
• Promass 80 PROFIBUS PA
– ID No.: 1528 (hex)
– Extended GSD file: EH3x1528.gsd
– Standard GSD file: EH3_1528.gsd
• Promass 83 PROFIBUS PA
– ID No.: 152A (hex)
– Extended GSD file: EH3x152A.gsd
– Standard GSD file: EH3_152A.gsd
Description of the function scope of compatibility:
Operating Instructions → 120.
• Cyclic data transmission
• Block model
• Description of the modules
PROFIBUS DP
Manufacturer ID0x11
Ident number0x156F
Profile version3.02
Device description files (GSD,
DTM, DD)
Supported functions• Identification & Maintenance
Information and files under:
• www.endress.com
On the product page for the device: Documents/Software → Device drivers
• www.profibus.org
Simplest device identification on the part of the control system and
nameplate
• PROFIBUS upload/download
Reading and writing parameters is up to ten times faster with PROFIBUS
upload/download
• Condensed status
Simplest and self-explanatory diagnostic information by categorizing
diagnostic messages that occur
Endress+Hauser23
Page 24
Proline Promass F 300
Configuration of the device
address
System integrationInformation regarding system integration: Operating Instructions → 120.
• DIP switches on the I/O electronics module
• Via operating tools (e.g. FieldCare)
• Cyclic data transmission
• Block model
• Description of the modules
EtherNet/IP
Protocol• The CIP Networks Library Volume 1: Common Industrial Protocol
• The CIP Networks Library Volume 2: EtherNet/IP Adaptation of CIP
Communication type• 10Base-T
• 100Base-TX
Device profileGeneric device (product type: 0x2B)
Manufacturer ID0x11
Device type ID0x103B
Baud ratesAutomatic ¹⁰⁄₁₀₀ Mbit with half-duplex and full-duplex detection
PolarityAuto-polarity for automatic correction of crossed TxD and RxD pairs
Supported CIP connectionsMax. 3 connections
Explicit connectionsMax. 6 connections
I/O connectionsMax. 6 connections (scanner)
Configuration options for
measuring device
Configuration of the EtherNet
interface
Configuration of the device
address
Device Level Ring (DLR)Yes
System integrationInformation regarding system integration: Operating Instructions
• DIP switches on the electronics module for IP addressing
• Manufacturer-specific software (FieldCare)
• Add-on Profile Level 3 for Rockwell Automation control systems
• Web browser
• Electronic Data Sheet (EDS) integrated in the measuring device
• Speed: 10 MBit, 100 MBit, auto (factory setting)
• Duplex: half-duplex, full-duplex, auto (factory setting)
• DIP switches on the electronics module for IP addressing (last octet)
• DHCP
• Manufacturer-specific software (FieldCare)
• Add-on Profile Level 3 for Rockwell Automation control systems
• Web browser
• EtherNet/IP tools, e.g. RSLinx (Rockwell Automation)
→ 120.
• Cyclic data transmission
• Block model
• Input and output groups
PROFINET
Protocol"Application layer protocol for decentral device periphery and distributed
automation", version 2.3
Communication type100 MBit/s
Conformity classConformance Class B
Netload ClassNetload Class II
Baud ratesAutomatic 100 Mbit/s with full-duplex detection
Cycle timesFrom 8 ms
PolarityAuto-polarity for automatic correction of crossed TxD and RxD pairs
Response times• Direct data access: typically 25 to 50 ms
• Auto-scan buffer (data range): typically 3 to 5 ms
Device typeSlave
Slave address range1 to 247
Broadcast address range0
Function codes• 03: Read holding register
• 04: Read input register
• 06: Write single registers
• 08: Diagnostics
• 16: Write multiple registers
• 23: Read/write multiple registers
Broadcast messagesSupported by the following function codes:
• 06: Write single registers
• 16: Write multiple registers
• 23: Read/write multiple registers
26Endress+Hauser
Page 27
Proline Promass F 300
Supported baud rate• 1200 BAUD
• 2400 BAUD
• 4800 BAUD
• 9600 BAUD
• 19200 BAUD
• 38400 BAUD
• 57600 BAUD
• 115200 BAUD
Data transfer mode• ASCII
• RTU
Data accessEach device parameter can be accessed via Modbus RS485.
For Modbus register information
Compatibility with
earlier model
System integrationInformation on system integration: Operating Instructions → 120.
If the device is replaced, the measuring device Promass 300 supports the
compatibility of the Modbus registers for the process variables and the
diagnostic information with the previous model Promass 83. It is not
necessary to change the engineering parameters in the automation system.
Description of the function scope of compatibility:
Terminal assignment of the remote display and operating module → 33.
Device plugs may not be used in hazardous areas!
Device plugs for fieldbus systems:
Order code for "Input; output 1"
• Option SA "FOUNDATION Fieldbus" → 28
• Option GA "PROFIBUS PA" → 28
• Option RA "PROFINET" → 29
• Option NA "EtherNet/IP" → 29
Device plug for connecting to the service interface:
Order code for "Accessory mounted"
option NB, adapter RJ45 M12 (service interface) → 30
Order code for "Input; output 1", option SA "FOUNDATION Fieldbus"
Order code forCable entry/connection → 32
"Electrical connection"23
M, 3, 4, 57/8" connector–
Order code for "Input; output 1", option GA "PROFIBUS PA"
Order code forCable entry/connection → 32
"Electrical connection"23
L, N, P, UConnector M12 × 1–
28Endress+Hauser
Page 29
Proline Promass F 300
1
2
4
3
1
2
4
3
Order code for "Input; output 1", option RA "PROFINET"
Order code forCable entry/connection → 32
"Electrical connection"23
L, N, P, UConnector M12 × 1–
1) 2)
1) 2)
1) 2)
R
, S
, T
, V
1) 2)
Connector M12 × 1Connector M12 × 1
1)Cannot be combined with an external WLAN antenna (order code for "Enclosed accessories", option P8) of
an RJ45 M12 adapter for the service interface (order code for "Accessories mounted", option NB) or of the
remote display and operating module DKX001.
2)Suitable for integrating the device in a ring topology.
Order code for "Input; output 1", option NA "EtherNet/IP"
Order code forCable entry/connection → 32
"Electrical connection"23
L, N, P, UConnector M12 × 1–
1) 2)
1) 2)
1) 2)
R
, S
, T
, V
1) 2)
Connector M12 × 1Connector M12 × 1
1)Cannot be combined with an external WLAN antenna (order code for "Enclosed accessories", option P8) of
an RJ45 M12 adapter for the service interface (order code for "Accessories mounted", option NB) or of the
remote display and operating module DKX001
2)Suitable for integrating the device in a ring topology.
Order code for "Accessory mounted", option NB "Adapter RJ45 M12 (service interface)"
Order codeCable entry/coupling → 32
"Accessory mounted"Cable entry
NBPlug M12 × 1–
Pin assignment, device plugFOUNDATION Fieldbus
PROFIBUS PA
Cable entry
2
3
PinAssignmentCodingPlug/socket
1+Signal +APlug
2-Signal –
3Grounding
4Not assigned
PinAssignmentCodingPlug/socket
1+PROFIBUS PA +APlug
2Grounding
3-PROFIBUS PA –
4Not assigned
Endress+Hauser29
Page 30
PROFINET
3
2
4
3
2
4
3
2
4
PinAssignment
1+TD +
2+RD +
3-TD –
4-RD –
A0032047
Recommended plug:
• Binder, series 763, part no. 99 3729 810 04
• Phoenix, part no. 1543223 SACC-M12MSD-4Q
• When using the device in a hazardous location, use a suitably certified plug.
CodingPlug/socket
DSocket
EtherNet/IP
PinAssignment
1+Tx
2+Rx
3-Tx
4-Rx
A0032047
CodingPlug/socket
DSocket
Proline Promass F 300
Supply voltage
Recommended plug:
• Binder, series 763, part no. 99 3729 810 04
• Phoenix, part no. 1543223 SACC-M12MSD-4Q
• When using the device in a hazardous location, use a suitably certified plug.
Service interface
Order code for "Accessories mounted", option NB: Adapter RJ45 M12 (service interface)
PinAssignment
1+Tx
2+Rx
3-Tx
4-Rx
A0032047
Recommended plug:
• Binder, series 763, part no. 99 3729 810 04
• Phoenix, part no. 1543223 SACC-M12MSD-4Q
• When using the device in a hazardous location, use a suitably certified plug.
Order code for
"Power supply"
Option DDC24 V±20%–
Option EAC100 to 240 V –15…+10%50/60 Hz
Option I
CodingPlug/socket
DSocket
terminal voltageFrequency range
DC24 V±20%–
AC100 to 240 V –15…+10%50/60 Hz
30Endress+Hauser
Page 31
Proline Promass F 300
Power consumptionTransmitter
Max. 10 W (active power)
Current consumptionTransmitter
• Max. 400 mA (24 V)
• Max. 200 mA (110 V, 50/60 Hz; 230 V, 50/60 Hz)
Power supply failure
Depending on the device version, the configuration is retained in the device memoryor in the
pluggable data memory (HistoROM DAT).
Endress+Hauser31
Page 32
Electrical connectionConnecting the transmitter
1
2
3
4
1
2
3
4
• Terminal assignment→ 27
• Device plugs available→ 28
1Terminal connection for supply voltage
2Terminal connection for signal transmission, input/output
3Terminal connection for signal transmission, input/output or terminal for network connection via service
interface (CDI-RJ45); Optional: terminal connection for external WLAN antenna or connection for remote
display and operating module DKX001
4Protective ground (PE)
An adapter for RJ45 and the M12 connector is optionally available:
Order code for "Accessories", option NB: "Adapter RJ45 M12 (service interface)"
The adapter connects the service interface (CDI-RJ45) to an M12 connector mounted in the
cable entry. Therefore the connection to the service interface can be established via an M12
connector without opening the device.
Network connection via service interface (CDI-RJ45) → 107
Proline Promass F 300
A0026781
Connecting in a ring topology
Device versions with EtherNet/IP and PROFINET communication protocols can be integrated into a
ring topology. The device is integrated via the terminal connection for signal transmission (output 1)
and the connection to the service interface (CDI-RJ45).
Integrate the transmitter into a ring topology:
• EtherNet/IP → 105
• PROFINET → 106
A0026781
1Terminal connection for supply voltage
2Terminal connection for signal transmission: PROFINET or EtherNet/IP (RJ45 connector)
3Connection to service interface (CDI-RJ45)
4Protective earth (PE)
If the device has additional input/outputs, these are routed via the cable entry for the
connection to the service interface (CDI-RJ45).
32Endress+Hauser
Page 33
Proline Promass F 300
1
3
81
Vcc
82
Gnd
83A84
B
81
Vcc
82
Gnd
83A84
B
2
4
5
4
4...20 mA
5
2
1
3
6
Connecting the remote display and operating module DKX001
The remote display and operating module DKX001 is available as an optional extra → 117.
• The remote display and operating module DKX001 is only available for the following housing
versions, order code for "Housing":
– Option A "Aluminum, coated"
– Option L "Cast, stainless"
• The measuring device is always supplied with a dummy cover when the remote display and
operating module DKX001 is ordered directly with the measuring device. Display or operation
at the transmitter is not possible in this case.
• If ordered subsequently, the remote display and operating module DKX001 may not be
connected at the same time as the existing measuring device display module. Only one
display or operation unit may be connected to the transmitter at any one time.
2Connection example for 4 to 20 mA HART current output (active)
1Automation system with current input (e.g. PLC)
2Cable shield: the cable shield must be grounded at both ends to comply with EMC requirements; observe cable
specifications → 41
3Connection for HART operating devices → 101
4Resistor for HART communication (≥ 250 Ω): observe maximum load → 15
5Analog display unit: observe maximum load → 15
6Transmitter
Endress+Hauser33
Page 34
Proline Promass F 300
2
3
4...20 mA
41
5
2
4...20 mA
4
1
2
3
3
6
5
A0028762
3Connection example for 4 to 20 mA HART current output (passive)
1Automation system with current input (e.g. PLC)
2Power supply
3Cable shield: the cable shield must be grounded at both ends to comply with EMC requirements; observe cable
4Connection example for HART input with a common negative (passive)
1Automation system with HART output (e.g. PLC)
2Active barrier for power supply (e.g. RN221N)
3Cable shield: the cable shield must be grounded at both ends to comply with EMC requirements; observe cable
specifications
4Analog display unit: observe maximum load
5Pressure transmitter (e.g. Cerabar M, Cerabar S): see requirements
6Transmitter
34Endress+Hauser
Page 35
Proline Promass F 300
2134
78
66
6
6
5
6
6
5
2
1
A
B
3
4
4
A
B
A
B
PROFIBUS PA
A0028768
5Connection example for PROFIBUS PA
1Control system (e.g. PLC)
2PROFIBUS PA segment coupler
3Cable shield: the cable shield must be grounded at both ends to comply with EMC requirements; observe cable
1Control system (e.g. PLC)
2Ethernet switch
3Observe cable specifications → 41
4Connecting cable between the two transmitters
5Transmitter
FOUNDATION Fieldbus
9Connection example for FOUNDATION Fieldbus
1Control system (e.g. PLC)
2Power Conditioner (FOUNDATION Fieldbus)
3Cable shield: the cable shield must be grounded at both ends to comply with EMC requirements; observe cable
10Connection example for Modbus RS485, non-hazardous area and Zone 2; Class I, Division 2
1Control system (e.g. PLC)
2Cable shield: the cable shield must be grounded at both ends to comply with EMC requirements; observe cable
specifications
3Distribution box
4Transmitter
Current output 4-20 mA
A0028758
11Connection example for 4-20 mA current output (active)
1Automation system with current input (e.g. PLC)
2Analog display unit: observe maximum load
3Transmitter
A0028759
12Connection example for 4-20 mA current output (passive)
1Automation system with current input (e.g. PLC)
2Active barrier for power supply (e.g. RN221N)
3Analog display unit: observe maximum load
4Transmitter
38Endress+Hauser
Page 39
Proline Promass F 300
1
2
3
12345
1
2
3
1
2
3
4
Pulse/frequency output
A0028761
13Connection example for pulse/frequency output (passive)
1Automation system with pulse/frequency input (e.g. PLC)
2Power supply
3Transmitter: Observe input values → 16
Switch output
14Connection example for switch output (passive)
1Automation system with switch input (e.g. PLC)
2Power supply
3Transmitter: Observe input values → 16
Double pulse output
15Connection example for double pulse output (active)
1Automation system with relay input (e.g. PLC)
2Power supply
3Transmitter: Observe input values → 18
Current input
A0028760
A0028915
18Connection example for 4 to 20 mA current input
1Power supply
2Terminal box
3External measuring device (for reading in pressure or temperature, for instance)
4Transmitter
40Endress+Hauser
Page 41
Proline Promass F 300
1
2
3
Status input
19Connection example for status input
1Automation system with status output (e.g. PLC)
2Power supply
3Transmitter
Potential equalizationRequirements
No special measures for potential equalization are required.
Please consider the following to ensure correct measurement:
• Same electrical potential for the fluid and sensor
• Company-internal grounding concepts
A0028764
terminals
Spring-loaded terminals: Suitable for strands and strands with ferrules.
Conductor cross-section 0.2 to 2.5 mm2 (24 to 12 AWG).
Cable entries
• Cable gland: M20 × 1.5 with cable ⌀ 6 to 12 mm (0.24 to 0.47 in)
• Thread for cable entry:
– NPT ½"
– G ½"
– M20
• Device plug for digital communication: M12
Only available for certain device versions → 28.
Cable specificationPermitted temperature range
• The installation guidelines that apply in the country of installation must be observed.
• The cables must be suitable for the minimum and maximum temperatures to be expected.
Power supply cable
Standard installation cable is sufficient.
Signal cable
Current output 4 to 20 mA HART
A shielded cable is recommended. Observe grounding concept of the plant.
PROFIBUS PA
Twisted, shielded two-wire cable. Cable type A is recommended .
For further information on planning and installing PROFIBUS networks see:
• Operating Instructions "PROFIBUS DP/PA: Guidelines for planning and commissioning"
(BA00034S)
• PNO Directive 2.092 "PROFIBUS PA User and Installation Guideline"
• IEC 61158-2 (MBP)
PROFIBUS DP
The IEC 61158 standard specifies two types of cable (A and B) for the bus line which can be used for
every transmission rate. Cable type A is recommended.
Endress+Hauser41
Page 42
Proline Promass F 300
Cable typeA
Characteristic impedance135 to 165 Ω at a measuring frequency of 3 to 20 MHz
Cable capacitance< 30 pF/m
Wire cross-section
Cable typeTwisted pairs
Loop resistance≤110 Ω/km
Signal dampingMax. 9 dB over the entire length of the cable cross-section
ShieldCopper braided shielding or braided shielding with foil shield. When grounding
For further information on planning and installing PROFIBUS networks see:
• Operating Instructions "PROFIBUS DP/PA: Guidelines for planning and commissioning"
(BA00034S)
• PNO Directive 2.092 "PROFIBUS PA User and Installation Guideline"
• IEC 61158-2 (MBP)
EtherNet/IP
The standard ANSI/TIA/EIA-568-B.2 Annex specifies CAT 5 as the minimum category for a cable
used for EtherNet/IP. CAT 5e and CAT 6 are recommended.
For more information on planning and installing EtherNet/IP networks, please refer to the
"Media Planning and Installation Manual. EtherNet/IP" of ODVA Organization
> 0.34 mm2 (22 AWG)
the cable shield, observe the grounding concept of the plant.
PROFINET
Standard IEC 61156-6 specifies CAT 5 as the minimum category for a cable used for PROFINET. CAT
5e and CAT 6 are recommended.
For more information on planning and installing PROFINET networks, see: "PROFINET Cabling
and Interconnection Technology", Guideline for PROFINET
FOUNDATION Fieldbus
Twisted, shielded two-wire cable.
For further information on planning and installing FOUNDATION Fieldbus networks see:
• Operating Instructions for "FOUNDATION Fieldbus Overview" (BA00013S)
• FOUNDATION Fieldbus Guideline
• IEC 61158-2 (MBP)
Modbus RS485
The EIA/TIA-485 standard specifies two types of cable (A and B) for the bus line which can be used
for every transmission rate. Cable type A is recommended.
Cable typeA
Characteristic impedance135 to 165 Ω at a measuring frequency of 3 to 20 MHz
Cable capacitance< 30 pF/m
Wire cross-section
Cable typeTwisted pairs
Loop resistance≤110 Ω/km
Signal dampingMax. 9 dB over the entire length of the cable cross-section
ShieldCopper braided shielding or braided shielding with foil shield. When grounding
> 0.34 mm2 (22 AWG)
the cable shield, observe the grounding concept of the plant.
Current output 0/4 to 20 mA
Standard installation cable is sufficient.
42Endress+Hauser
Page 43
Proline Promass F 300
Pulse/frequency/switch output
Standard installation cable is sufficient.
Double pulse output
Standard installation cable is sufficient.
Relay output
Standard installation cable is sufficient.
Current input 0/4 to 20 mA
Standard installation cable is sufficient.
Status input
Standard installation cable is sufficient.
Connecting cable for transmitter - remote display and operating module DKX001
Standard cable
A standard cable can be used as the connecting cable.
Standard cable4 cores (2 pairs); pair-stranded with common shield
The outputs have the following base accuracy specifications.
Current output
Accuracy±5 µA
Pulse/frequency output
o.r. = of reading
Repeatability
AccuracyMax. ±50 ppm o.r. (over the entire ambient temperature range)
o.r. = of reading; 1 g/cm3 = 1 kg/l; T = medium temperature
Base repeatability
Design fundamentals → 48
Mass flow and volume flow (liquids)
±0.025 % o.r. (PremiumCal, for mass flow)
±0.05 % o.r.
Mass flow (cryogenic liquids)
Order code for "Measuring tube material", option LA
±0.175 % % o.r.
Mass flow (gases)
±0.25 % o.r.
Density (liquids)
±0.00025 g/cm
3
Density (cryogenic liquids)
Order code for "Measuring tube material", option LA
±0.025 g/cm
3
Temperature
±0.25 °C ± 0.0025 · T °C (±0.45 °F ± 0.0015 · (T–32) °F)
Response time
Influence of ambient
The response time depends on the configuration (damping).
Current output
temperature
Temperature coefficientMax. 1 μA/°C
46Endress+Hauser
Page 47
Proline Promass F 300
3
[°C]
0- 0880 160 240
[° ]F
- 050
100
200
300
0
2
4
6
8
10
2
1
12
14
16
18
320 400 480 560
640
Pulse/frequency output
Temperature coefficientNo additional effect. Included in accuracy.
Influence of medium
temperature
Mass flow and volume flow
o.f.s. = of full scale value
When there is a difference between the temperature for zero point adjustment and the process
temperature, the additional measured error of the sensor is typically ±0.0002 % o.f.s./°C (±0.0001 %
o. f.s./°F).
The effect is reduced if zero point adjustment is performed at process temperature.
Density
When there is a difference between the density calibration temperature and the process
temperature, the typical measured error of the sensor is
±0.00005 g/cm3 /°C (±0.000025 g/cm3 /°F). Field density calibration is possible.
Wide-range density specification (special density calibration)
If the process temperature is outside the valid range (→ 44) the measured error is
±0.00005 g/cm3 /°C (±0.000025 g/cm3 /°F)
Influence of medium
pressure
Endress+Hauser47
A0027453
1Field density calibration, for example at +20 °C (+68 °F)
2Special density calibration
Temperature
±0.005 · T °C (± 0.005 · (T – 32) °F)
The table below shows the effect on accuracy of mass flow due to a difference between calibration
pressure and process pressure.
o.r. = of reading
It is possible to compensate for the effect by:
• Reading in the current pressure measured value via the current input.
• Specifying a fixed value for the pressure in the device parameters.
Operating Instructions→ 120.
DN[% o.r./bar][% o.r./psi]
[mm][in]
83/8no influence
15½no influence
Page 48
Proline Promass F 300
ZeroPoint
BaseAccu
⋅ 100
³
± BaseAccu
ZeroPoint
BaseAccu
⋅ 100
<
ZeroPoint
MeasValue
⋅ 100
±
½ ⋅ ZeroPoint
BaseRepeat
⋅ 100
³
½ ⋅ ZeroPoint
BaseRepeat
⋅ 100
<
ZeroPoint
MeasValue
⋅ 100
½ ⋅±
E [%]
0102030405060
0
0.5
1.0
1.5
2.0
2.5
708090100
Q [%]
DN[% o.r./bar][% o.r./psi]
[mm][in]
251no influence
401½–0.003–0.0002
502–0.008–0.0006
803–0.009–0.0006
1004–0.007–0.0005
1506–0.009–0.0006
25010–0.009–0.0006
Design fundamentals
o.r. = of reading, o.f.s. = of full scale value
BaseAccu = base accuracy in % o.r., BaseRepeat = base repeatability in % o.r.
MeasValue = measured value; ZeroPoint = zero point stability
Calculation of the maximum measured error as a function of the flow rate
Flow rateMaximum measured error in % o.r.
A0021332
A0021333
Calculation of the maximum repeatability as a function of the flow rate
Flow rateMaximum repeatability in % o.r.
A0021335
A0021336
A0021339
A0021334
A0021340
A0021337
Example for maximum measured error
EMaximum measured error in % o.r. (example with PremiumCal)
QFlow rate in % of maximum full scale value
Installation
No special measures such as supports etc. are necessary. External forces are absorbed by the
construction of the device.
48Endress+Hauser
A0028808
Page 49
Proline Promass F 300
1
2
3
4
5
Mounting location
A0028772
To prevent measuring errors arising from accumulation of gas bubbles in the measuring tube, avoid
the following mounting locations in the pipe:
• Highest point of a pipeline.
• Directly upstream of a free pipe outlet in a down pipe.
Installation in down pipes
However, the following installation suggestion allows for installation in an open vertical pipeline.
Pipe restrictions or the use of an orifice with a smaller cross-section than the nominal diameter
prevent the sensor running empty while measurement is in progress.
A0028773
20Installation in a down pipe (e.g. for batching applications)
1Supply tank
2Sensor
3Orifice plate, pipe restriction
4Valve
5Batching tank
DNØ orifice plate, pipe restriction
[mm][in][mm][in]
8³⁄₈60.24
15½100.40
251140.55
401½220.87
502281.10
803501.97
1004652.60
Endress+Hauser49
Page 50
Proline Promass F 300
12
DNØ orifice plate, pipe restriction
[mm][in][mm][in]
1506903.54
250101505.91
Orientation
The direction of the arrow on the sensor nameplate helps you to install the sensor according to the
flow direction (direction of medium flow through the piping).
OrientationRecommendation
AVertical orientation
A0015591
BHorizontal orientation, transmitter at
top
→ 21, 50
A0015589
CHorizontal orientation, transmitter at
bottom
→ 21, 50
A0015590
DHorizontal orientation, transmitter at
side
A0015592
1)Applications with low process temperatures may decrease the ambient temperature. To maintain the
minimum ambient temperature for the transmitter, this orientation is recommended.
2)Applications with high process temperatures may increase the ambient temperature. To maintain the
maximum ambient temperature for the transmitter, this orientation is recommended.
1)
Exceptions:
2)
Exceptions:
Inlet and outlet runs
Special mounting
instructions
If a sensor is installed horizontally with a curved measuring tube, match the position of the sensor to
the fluid properties.
A0028774
21Orientation of sensor with curved measuring tube
1Avoid this orientation for fluids with entrained solids: Risk of solids accumulating.
2Avoid this orientation for outgassing fluids: Risk of gas accumulating.
No special precautions need to be taken for fittings which create turbulence, such as valves, elbows
or T-pieces, as long as no cavitation occurs → 60.
Rupture disk
Information that is relevant to the process: → 60.
50Endress+Hauser
Page 51
Proline Promass F 300
RUPTURE DISK
i
2
3
1
DN 8 ( ")...150 (6")!⁄#DN 250 (10")
2
3
WARNING
L
Danger from medium escaping!
Medium escaping under pressure can cause injury or material damage.
Take precautions to prevent danger to persons and damage if the rupture disk is actuated.
‣
Observe information on the rupture disk sticker.
‣
Make sure that the function and operation of the rupture disk is not impeded through the
‣
installation of the device.
Do not use a heating jacket.
‣
Do not remove or damage the rupture disk.
‣
The position of the rupture disk is indicated on a sticker beside it.
The transportation guard must be removed.
The existing connecting nozzles are not intended for the purpose of rinsing or pressure monitoring,
but instead serve as the mounting location for the rupture disk.
In the event of a failure of the rupture disk, a discharge device can be screwed onto the internal
thread of the rupture disk in order to drain off any escaping medium.
1Rupture disk label
2Rupture disk with 1/2" NPT internal thread with 1" width across flat
3Transport protection
For information on the dimensions: see the "Mechanical construction" section (accessories)
Zero point adjustment
All measuring devices are calibrated in accordance with state-of-the-art technology. Calibration
takes place under reference conditions→ 44. Therefore, a zero point adjustment in the field is
generally not required.
Experience shows that zero point adjustment is advisable only in special cases:
• To achieve maximum measuring accuracy even with low flow rates.
• Under extreme process or operating conditions (e.g. very high process temperatures or very highviscosity fluids).
A0028903
Endress+Hauser51
Page 52
Environment
Proline Promass F 300
Ambient temperature range
Storage temperature
Climate class
Degree of protection
Measuring device• –40 to +60 °C (–40 to +140 °F)
• Order code for "Test, certificate", option JP:
–50 to +60 °C (–58 to +140 °F)
Readability of the local
display
Dependency of ambient temperature on medium temperature→ 53
If operating outdoors:
‣
Avoid direct sunlight, particularly in warm climatic regions.
You can order a weather protection cover from Endress+Hauser. → 117.
–50 to +80 °C (–58 to +176 °F)
DIN EN 60068-2-38 (test Z/AD)
Measuring device
• As standard: IP66/67, type 4X enclosure
• When housing is open: IP20, type 1 enclosure
• Display module: IP20, type 1 enclosure
• With the order code for "Sensor options", option CM: IP69 can also be ordered
External WLAN antenna
IP67
–20 to +60 °C (–4 to +140 °F)
The readability of the display may be impaired at temperatures outside the
temperature range.
Vibration resistance
Shock resistance
Shock resistance
Interior cleaning
Electromagnetic
compatibility (EMC)
• Oscillation, sinusoidal, following IEC 60068-2-6
– 2 to 8.4 Hz, 3.5 mm peak
– 8.4 to 2000 Hz, 1 g peak
• Oscillation, broadband noise following IEC 60068-2-64
– 10 to 200 Hz, 0.003 g2/Hz
– 200 to 2000 Hz, 0.001 g2/Hz
– Total: 1.54 g rms
Shock, half-sine according to IEC 60068-2-27
6 ms 50 g
Shock due to rough handling following IEC 60068-2-31
• Cleaning in place (CIP)
• Sterilization in place (SIP)
Options
• Oil- and grease-free version for wetted parts, without declaration
Order code for "Service", option HA
• Oil- and grease-free version for wetted parts as per IEC/TR 60877-2.0 and BOC 50000810-4, with
declaration
Order code for "Service", option HB
• As per IEC/EN 61326 and NAMUR Recommendation 21 (NE 21)
• Device version with PROFIBUS DP: Complies with emission limits for industry as per EN 50170
Volume 2, IEC 61784
The following applies for PROFIBUS DP: If baud rates > 1.5 MBaud, an EMC cable entry must be
used and the cable shield must continue as far as the terminal wherever possible.
Details are provided in the Declaration of Conformity.
52Endress+Hauser
Page 53
Proline Promass F 300
T
a
T
m
B
A
Medium temperature range
Process
Standard version–50 to +150 °C (–58 to +302 °F)Order code for "Measuring tube
mat., wetted surface", option HA,
SA, SB, SC
Extended temperature version–50 to +240 °C (–58 to +464 °F)Order code for "Measuring tube
mat., wetted surface", option SD, SE,
SF, TH
High-temperature version–50 to +350 °C (–58 to +662 °F)For nominal diameters DN 25 (1"),
DN 50 (2") and DN 80 (3")
Order code for "Measuring tube
mat., wetted surface", option TT, TU
Low-temperature version–196 to +150 °C (–320 to +302 °F)
NOTICE
Material fatigue due to excessive
temperature difference!
Maximum temperature
‣
difference of media used: 300 K
Order code for "Measuring tube
mat., wetted surface", option LA
Dependency of ambient temperature on medium temperature
22Exemplary representation, values in the table below.
TaAmbient temperature
TmMedium temperature
AMaximum permitted medium temperature Tm at T
require a reduced ambient temperature T
BMaximum permitted ambient temperature Ta for the maximum specified medium temperature Tm of the
sensor
a
= 60 °C (140 °F); higher medium temperatures T
a max
A0031121
m
Values for devices used in the hazardous area:
Separate Ex documentation (XA) for the device → 121.
Endress+Hauser53
Page 54
Proline Promass F 300
PN100
PN63
PN40
PN16
0
10
20
30
40
50
60
70
80
90
100
0
200
400
600
800
[bar][psi]
-320
64080160
240
320400480560
[°C]
[°F]
35030025020015010050
0
-200
Not insulatedInsulated
ABAB
VersionT
Standard version60 °C (140 °F) 150 °C (302 °F)––60 °C (140 °F) 110 °C (230 °F) 55 °C (131 °F)150 °C
Extended
temperature
version
High-temperature
version
a
60 °C (140 °F) 170 °C (338 °F) 55 °C (131 °F)240 °C
60 °C (140 °F) 350 °C (662 °F)––60 °C (140 °F) 350 °C (662 °F)––
T
m
T
a
T
m
(464 °F)
T
a
60 °C (140 °F) 110 °C (230 °F) 50 °C (122 °F)240 °C
T
m
T
a
T
m
(302 °F)
(464 °F)
Density
Pressure-temperature curves
0 to 5000 kg/m3 (0 to 312 lb/cf)
The following pressure/temperature diagrams apply to all pressure-bearing parts of the device and
not just the process connection. The diagrams show the maximum permissible medium pressure
depending on the specific medium temperature.
• Pressure-temperature curves with temperature range +151 to +240 °C (+304 to +464 °F)
exclusively for extended temperature version of measuring devices.
• Pressure-temperature curves with temperature range +241 to +350 °C (+466 to +662 °F)
exclusively for high temperature version of measuring devices.
• Pressure-temperature curves with temperature range –196 to +150 °C (–320 to +302 °F)
exclusively for low temperature version of measuring devices.
Flange according to EN 1092-1 (DIN 2501)
23With flange material 1.4404 (F316/F316L), Alloy C22
54Endress+Hauser
A0032852-EN
Page 55
Proline Promass F 300
Class 600
Class 300
Class 150
0
10
20
30
40
50
60
70
80
90
100
0
200
400
600
800
[bar][psi]
-320
64080160
240
320400480560
[°C]
[°F]
35030025020015010050
0
-200
Class 600
Class 300
Class 150
-80
640080160
240
320400480560
0
10
20
30
40
50
60
70
80
90
100
0
200
400
600
800
[°C]
[°F]
[bar][psi]
35030025020015010050
0
-50
Flange according to ASME B16.5
A0032853-EN
24With flange material 1.4404 (F316/F316L)
A0028780-EN
25With flange material Alloy C22
Endress+Hauser55
Page 56
Flange JIS B2220
63K
40K
20K
10K
0
10
20
30
40
50
60
70
0
200
400
600
800
[bar][psi]
-320
64080160
240
320400480560
[°C]
[°F]
35030025020015010050
0
-200
DN8…40
≥ DN50
-80
080160
240
320400480
0
10
20
30
40
0
[°C]
[°F]
[bar][psi]
25020015010050
0
-50
50
PN40
-80
080160
240
320400480
0
10
20
30
40
0
[°C]
[°F]
[bar][psi]
25020015010050
0
-50
50
26With flange material 1.4404 (F316/F316L), Alloy C22
Flange DIN 11864-2 Form A
Proline Promass F 300
A0032851-EN
27With connection material 1.4404 (316/316L)
Lap joint flange according to EN 1092-1 (DIN 2501)
28With flange material 1.4301 (F304); wetted parts Alloy C22
56Endress+Hauser
A0028782-EN
A0028784-EN
Page 57
Proline Promass F 300
Class 150
Class 600
Class 300
-80
080160
240
320400480
0
10
20
30
40
50
60
70
80
90
100
0
200
400
600
800
[°C]
[°F]
[bar][psi]
25020015010050
0
-50
20K
-80
080160
240
320400480
0
10
20
30
0
200
400
[°C]
[°F]
[bar]
[psi]
25020015010050
0
-50
≥ DN50
DN8…40
-80
080160
240
320400480
0
10
20
30
40
0
[°C]
[°F]
[bar][psi]
25020015010050
0
-50
50
Lap joint flange according to ASME B16.5
A0028785-EN
29With flange material 1.4301 (F304); wetted parts Alloy C22
Lap joint flange JIS B2220
A0028786-EN
30With flange material 1.4301 (F304); wetted parts Alloy C22
Thread DIN 11851
Endress+Hauser57
31With connection material 1.4404 (316/316L)
DIN 11851 allows for applications up to +140 °C (+284 °F) if suitable sealing materials are used.
Please take this into account when selecting seals and counterparts, as these components can limit
the pressure and temperature range.
A0028794-EN
Page 58
Thread DIN 11864-1 Form A
DN8…40
≥ DN50
-80
080160
240
320400480
0
10
20
30
40
0
[°C]
[°F]
[bar][psi]
25020015010050
0
-50
50
-80
080160
240
320400480
0
10
20
0
[°C]
[°F]
[bar][psi]
25020015010050
0
-50
-80
080160
240
320400480
0
10
20
0
[°C]
[°F]
[bar][psi]
25020015010050
0
-50
32With connection material 1.4404 (316/316L)
Thread ISO 2853
Proline Promass F 300
A0028798-EN
A0028799-EN
33With connection material 1.4404 (316/316L)
Thread SMS 1145
A0028800-EN
34With connection material 1.4404 (316/316L)
SMS 1145 allows for applications up to 16 bar (232 psi) if suitable sealing materials are used. Please
take this into account when selecting seals and counterparts, as these components can limit the
pressure and temperature range.
58Endress+Hauser
Page 59
Proline Promass F 300
60
70
80
90
100
800
[bar][psi]
-80
080160
240
320400480
[°C]
[°F]
25020015010050
0
-50
-80
080160
240
320400480
0
10
20
0
[°C]
[°F]
[bar][psi]
25020015010050
0
-50
VCO
A0028801-EN
35With connection material 1.4404 (316/316L)
Tri-Clamp
The clamp connections are suitable up to a maximum pressure of 16 bar (232 psi). Please observe
the operating limits of the clamp and seal used as they can be over 16 bar (232 psi). The clamp and
seal are not included in the scope of supply.
Sensor housing
For standard versions with the temperature range –50 to +150 °C (–58 to +302 °F), the sensor
housing is filled with dry nitrogen gas and protects the electronics and mechanics inside.
For all other temperature versions the sensor housing is filled with dry inert gas.
If a measuring tube fails (e.g. due to process characteristics like corrosive or abrasive fluids), the
fluid will initially be contained by the sensor housing.
In the event of a tube failure, the pressure level inside the sensor housing will rise according to the
operating process pressure. If the user judges that the sensor housing pressure rating/burst pressure
does not provide an adequate safety margin, the device can be fitted with a rupture disk. This
prevents excessively high pressure from forming inside the sensor housing. Therefore, the use of a
rupture disk is strongly recommended in applications involving high gas pressures, and particularly
in applications in which the process pressure is greater than 2/3 of the sensor housing burst
pressure.
If there is a need to drain the leaking medium into a discharge device, the sensor should be fitted
with a rupture disk. Connect the discharge to the additional threaded connection → 82.
If the sensor is to be purged with gas (gas detection), it should be equipped with purge connections.
Do not open the purge connections unless the containment can be filled immediately with a
dry, inert gas. Use only low pressure to purge. Maximum pressure: 5 bar (72.5 psi).
Endress+Hauser59
Sensor housing nominal pressure rating and burst pressure
The following sensor housing nominal pressure ratings/burst pressures are only valid for standard
devices and/or devices equipped with closed purge connections (not opened/as delivered).
A0032216-EN
Page 60
Proline Promass F 300
If a device fitted with purge connections (order code for "Sensor option", option CH "Purge
connection") is connected to the purge system, the maximum nominal pressure is determined by the
purge system itself or by the device, depending on which component has the lower nominal pressure
classification.
If the device is fitted with a rupture disk (order code for "Sensor option", option CA "Rupture disk"), the
rupture disk trigger pressure is decisive for the maximum nominal pressure .
The sensor housing burst pressure refers to a typical internal pressure which is reached prior to
mechanical failure of the sensor housing and which was determined during type testing. The
corresponding type test declaration can be ordered with the device (order code for "Additional
approval", option LN "Sensor housing burst pressure, type test").
Rupture disk
DNSensor housing nominal
pressure
(designed with a safety factor
≥ 4)
[mm][in][bar][psi][bar][psi]
8³⁄₈405802553698
15½405802002900
251405802804060
401½405801802610
502405801952828
803253621051522
100416232851232
150616232801160
250101014557826
Sensor housing burst pressure
For information on the dimensions: see the "Mechanical construction" section
To increase the level of safety, a device version with a rupture disk with a trigger pressure of
10 to 15 bar (145 to 217.5 psi)can be used (order code for "Sensor option", option CA "rupture disk").
The use of rupture disks cannot be combined with the separately available heating jacket.
Information on the dimensions of the rupture disk: → 82
Flow limit
Select the nominal diameter by optimizing between the required flow range and permissible pressure
loss.
For an overview of the full scale values for the measuring range, see the "Measuring range"
section → 11
• The minimum recommended full scale value is approx. 1/20 of the maximum full scale value
• In most applications, 20 to 50 % of the maximum full scale value can be considered ideal
• A low full scale value must be selected for abrasive media (such as liquids with entrained solids):
flow velocity < 1 m/s (< 3 ft/s).
• For gas measurement the following rules apply:
– The flow velocity in the measuring tubes should not exceed half the sound velocity (0.5 Mach).
– The maximum mass flow depends on the density of the gas: formula → 11
To calculate the flow limit, use the Applicator sizing tool → 119
Pressure loss
To calculate the pressure loss, use the Applicator sizing tool → 119
Promass F with reduced pressure loss: order code for "Sensor option", option CE "Reduced pressure
loss"
System pressure
It is important that cavitation does not occur, or that gases entrained in the liquids do not outgas.
This is prevented by means of a sufficiently high system pressure.
60Endress+Hauser
Page 61
Proline Promass F 300
For this reason, the following mounting locations are recommended:
• At the lowest point in a vertical pipe
• Downstream from pumps (no danger of vacuum)
A0028777
Thermal insulation
In the case of some fluids, it is important to keep the heat radiated from the sensor to the
transmitter to a low level. A wide range of materials can be used for the required insulation.
The following device versions are recommended for versions with thermal insulation:
• Version with extended neck for insulation:
Order code for "Sensor option", option CG with an extended neck length of 105 mm (4.13 in).
• Extended temperature version:
Order code for "Measuring tube material", option SD, SE, SF or TH with an extended neck length of
105 mm (4.13 in).
• High-temperature version:
Order code for "Measuring tube material", option TT or TU with an extended neck length of
142 mm (5.59 in).
NOTICE
Electronics overheating on account of thermal insulation!
Maximum permissible temperature at the lower end of the transmitter housing: 80 °C (176 °F)
‣
Thermal insulation with extended neck free: We recommend that you do not insulate the
‣
extended neck in order to ensure optimum dissipation of heat.
A0034391
36Thermal insulation with extended neck free
Low-temperature version: It is generally not necessary to insulate the transmitter housing . If
insulation is provided, the rules that apply are the same as those for thermal insulation.
Heating
Some fluids require suitable measures to avoid loss of heat at the sensor.
Heating options
• Electrical heating, e.g. with electric band heaters
• Via pipes carrying hot water or steam
• Via heating jackets
Heating jackets for the sensors can be ordered as accessories from Endress+Hauser .→ 118
Endress+Hauser61
Page 62
Proline Promass F 300
NOTICE
Danger of overheating when heating
Ensure that the temperature at the lower end of the transmitter housing does not exceed
‣
80 °C (176 °F).
Ensure that sufficient convection takes place at the transmitter neck.
‣
Ensure that a sufficiently large area of the transmitted neck remains exposed. The uncovered
‣
part serves as a radiator and protects the electronics from overheating and excessive cooling.
If using in potentially explosive atmospheres, observe the information in the device-specific Ex
‣
documentation. For detailed information on the temperature tables, see the separate document
entitled "Safety Instructions" (XA) for the device.
Vibrations
The high oscillation frequency of the measuring tubes ensures that the correct operation of the
measuring system is not influenced by plant vibrations.
62Endress+Hauser
Page 63
Proline Promass F 300
Custody transfer measurement
The measuring device is optionally tested in accordance with OIML R117/R81 and has an
EU type evaluation certificate which authorizes the use in EU type-examination certificates according
to Measuring Instruments Directive 2014/32/EU for service subject to legal metrological control
("custody transfer") for liquids other than water and cryogenic liquids (Annex VII).
The permitted fluid temperature in these applications is –200 to +90 °C (–328 to +194 °F).
The measuring device is optionally tested in accordance with OIML R137 and has an EU typeexamination certificate according to Measuring Instruments Directive 2014/32/EU for service as a
gas meter subject to legal metrological control ("custody transfer") (Annex IV). The permitted fluid
temperature in these applications is –25 to +55 °C (–13 to +131 °F).
The device is used with a legally controlled totalizer on the local display and optionally with legally
controlled outputs.
Measuring devices subject to legal metrological control totalize in both directions, i.e. all the outputs
consider flow components in the positive (forward) and negative (reverse) flow direction.
Generally a measuring device subject to legal metrological control is secured against tampering by
seals on the transmitter or sensor. These seals may normally only be opened by a representative of
the competent authority for legal metrology controls.
After putting the device into circulation or after sealing the device, operation is only possible to a
limited extent.
Detailed ordering information is available from your local Endress+Hauser sales center for national
approvals, which are based on the OIML certificates, for applications with liquids other than water,
cryogenic liquids or gases.
Endress+Hauser63
Page 64
Mechanical construction
D
L
E
F
A
CB
K
G
H
I
M
Dimensions in SI unitsCompact version
Proline Promass F 300
A0033786
Order code for "Housing", option A "Aluminum, coated"
1)Depending on the cable gland used: values up to + 30 mm
2)With order code for "Sensor option", option CG or order code for "Measuring tube material", option SD, SE,
SF, TH, LA: values +70 mm
3)With order code for "Measuring tube material", option TT, TU: values +104 mm
4)Depends on the process connection in question
2) 3)
FGHIKLM
4)
70
4)
70
4)
70
4)
79
4)
99
4)
139
4)
176
4)
218
4)
305
Endress+Hauser65
Page 66
Flange connections
C
D
A
B
E
L
Fixed flange EN 1092-1, ASME B16.5, JIS B2220
Length tolerance for dimension L in mm:
• DN ≤ 100: +1.5 / –2.0
• DN ≥ 125: +3.5
Proline Promass F 300
A0015621
Flange according to EN 1092-1 (DIN 2501): PN16
1.4404 (F316/F316L): order code for "Process connection", option D1S
Alloy C22: order code for "Process connection", option D1C
Flange with groove according to EN 1092-1 Form D (DIN 2512N): PN16
1.4404 (F316/F316L): order code for "Process connection", option D5S
Alloy C22: order code for "Process connection", option D5C
DN
[mm]
1002201808 × ⌀1820107.11127/1400
1502852408 × ⌀2222159.31330/1700
25040535512 × ⌀2626260.41775
Surface roughness (flange): EN 1092-1 Form B1 (DIN 2526 Form C), Ra 3.2 to 12.5 µm
1)Installation length in accordance with NAMUR recommendation NE 132 optionally available (order code
for "Process connection", option D1N or D5N (with groove))
Flange according to EN 1092-1 (DIN 2501): PN16 with reduction in nominal diameter
1.4404 (F316/F316L
DN
[mm]
10080DHS2201808 × ⌀ 1820107.1874
150100DJS2852408 × ⌀ 2222159.31167
200150DLS34029512 × ⌀ 2224206.51461
Surface roughness (flange): EN 1092-1 Form B1 (DIN 2526 Form C), Ra 3.2 to 12.5 µm
A
[mm]
reduction
to DN
[mm]
B
[mm]
Order code for
"Process
connection",
option
C
[mm]
A
[mm]B[mm]
D
[mm]
C
[mm]
E
[mm]
D
[mm]E[mm]L[mm]
L
[mm]
1)
1)
66Endress+Hauser
Page 67
Proline Promass F 300
Flange according to EN 1092-1 (DIN 2501): PN 40
1.4404 (F316/F316L): order code for "Process connection", option D2S
Alloy C22: order code for "Process connection", option D2C
Flange with groove according to EN 1092-1 Form D (DIN 2512N): PN 40
1.4404 (F316/F316L): order code for "Process connection", option D6S
Alloy C22: order code for "Process connection", option D6C
DN
[mm]
1)
8
A
[mm]
B
[mm]
C
[mm]
D
[mm]
E
[mm]
95654 × ⌀141617.3370/510
1595654 × ⌀141617.3404/510
25115854 × ⌀141828.5440/600
401501104 × ⌀181843.1550
501651254 × ⌀182054.5715/715
802001608 × ⌀182482.5840/915
1002351908 × ⌀2224107.11127
1503002508 × ⌀2628159.31370
25045038512 × ⌀3338258.81845
Surface roughness (flange): EN 1092-1 Form B1 (DIN 2526 Form C), Ra 3.2 to 12.5 µm
L
[mm]
2)
2)
2)
2)
2)
1)DN 8 with DN 15 flanges as standard
2)Installation length in accordance with NAMUR recommendation NE 132 optionally available (order code
for "Process connection", option D2N or D6N (with groove))
Flange according to EN 1092-1 (DIN 2501): PN 40 (with DN 25 flanges)
1.4404 (F316/F316L): order code for "Process connection", option R2S
DN
[mm]
A
[mm]
B
[mm]
C
[mm]
D
[mm]
E
[mm]
L
[mm]
8115854 × ⌀141828.5440
15115854 × ⌀141828.5440
Surface roughness (flange): EN 1092-1 Form B1 (DIN 2526 Form C), Ra 3.2 to 12.5 µm
Flange according to EN 1092-1 (DIN 2501): PN 40 with reduction in nominal diameter
1.4404 (F316/F316L)
DN
[mm]
reduction
to DN
[mm]
Order code for
"Process
connection",
A
[mm]B[mm]
C
[mm]
D
[mm]E[mm]L[mm]
option
5040DFS1651254 × ⌀ 182054.5555
8050DGS2001608 × ⌀ 182482.5840
10080DIS2351908 × ⌀ 2224107.1874
150100DKS3002508 × ⌀ 2628159.31167
200150DMS37532012 × ⌀ 3034206.51461
Surface roughness (flange): EN 1092-1 Form B1 (DIN 2526 Form C), Ra 3.2 to 12.5 µm
Endress+Hauser67
Page 68
Proline Promass F 300
Flange according to EN 1092-1 (DIN 2501): PN 63
1.4404 (F316/F316L): order code for "Process connection", option D3S
Alloy C22: order code for "Process connection", option D3C
Flange with groove according to EN 1092-1 Form D (DIN 2512N): PN 63
1.4404 (F316/F316L): order code for "Process connection", option D7S
Alloy C22: order code for "Process connection", option D7C
DN
[mm]
501801354 × ⌀222654.5724
802151708 × ⌀222881.7875
1002502008 × ⌀2630106.31127
1503452808 × ⌀3336157.11410
25047040012 × ⌀3646255.41885
Surface roughness (flange):
EN 1092-1 Form B1 (DIN 2526 Form C), Ra 3.2 to 12.5 µm EN 1092-1 Form B2 (DIN 2526 Form E), Ra
0.8 to 3.2 µm
Flange according to EN 1092-1 (DIN 2501): PN 100
1.4404 (F316/F316L): order code for "Process connection", option D4S
Alloy C22: order code for "Process connection", option D4C
Flange with groove according to EN 1092-1 Form D (DIN 2512N): PN 100
1.4404 (F316/F316L): order code for "Process connection", option D8S
Alloy C22: order code for "Process connection", option D8C
DN
[mm]
1)
8
15105754 × ⌀142017.3420
251401004 × ⌀182428.5470
401701254 × ⌀222642.5590
501951454 × ⌀262853.9740
802301808 × ⌀263280.9885
1002652108 × ⌀3036104.31127
15035529012 × ⌀3344154.01450
Surface roughness (flange): EN 1092-1 Form B2 (DIN 2526 Form E), Ra 0.8 to 3.2 µm
A
[mm]
A
[mm]
105754 × ⌀142017.3400
B
[mm]
B
[mm]
C
[mm]
C
[mm]
D
[mm]
D
[mm]
E
[mm]
E
[mm]
[mm]
[mm]
L
L
1)DN 8 with DN 15 flanges as standard
Flange according to EN 1092-1 (DIN 2501): PN 100
Alloy C22: order code for "Process connection", option D4C
Flange with groove according to EN 1092-1 Form D (DIN 2512N): PN 100
Alloy C22: order code for "Process connection", option D8C
DN
[mm]
25050543012 × ⌀3960248.01949
Surface roughness (flange): EN 1092-1 Form B2 (DIN 2526 Form E), Ra 0.8 to 3.2 µm
A
[mm]
B
[mm]
C
[mm]
D
[mm]
E
[mm]
L
[mm]
68Endress+Hauser
Page 69
Proline Promass F 300
Flange according to ASME B16.5: Class 150
1.4404 (F316/F316L): order code for "Process connection", option AAS
Alloy C22: order code for "Process connection", option AAC
DN
[mm]
1)
8
A
[mm]
B
[mm]
C
[mm]
D
[mm]
E
[mm]
9060.34 × ⌀15.711.215.7370
159060.34 × ⌀15.711.215.7404
2511079.44 × ⌀15.714.226.7440
4012598.44 × ⌀15.717.540.9550
50150120.74 × ⌀19.119.152.6715
80190152.44 × ⌀19.123.978.0840
100230190.58 × ⌀19.123.9102.41127
150280241.38 × ⌀22.425.4154.21398
25040536212 × ⌀25.430.2254.51832
Surface roughness (flange): Ra 3.2 to 6.3 µm
1)DN 8 with DN 15 flanges as standard
L
[mm]
Flange according to ASME B16.5: Class 150 with reduction in nominal diameter
1.4404 (F316/F316L)
DN
[mm]
reduction
to DN
[mm]
Order code for
"Process
connection",
A
[mm]B[mm]
C
[mm]
D
[mm]E[mm]L[mm]
option
5040AHS150120.74 × ⌀ 19.119.152.6550
8050AJS190152.44 × ⌀ 19.123.978.0720
10080ALS230190.58 × ⌀ 19.123.9102.4874
150100ANS280241.38 × ⌀ 22.425.4154.21167
200150APS345298.58 × ⌀ 22.429202.71461
Surface roughness (flange): Ra 3.2 to 6.3 µm
Flange according to ASME B16.5: Class 300
1.4404 (F316/F316L): order code for "Process connection", option ABS
Alloy C22: order code for "Process connection", option ABC
DN
[mm]
1)
8
A
[mm]
B
[mm]
C
[mm]
D
[mm]
E
[mm]
9566.74 × ⌀15.714.215.7370
[mm]
159566.74 × ⌀15.714.215.7404
2512588.94 × ⌀19.117.526.7440
40155114.34 × ⌀22.320.640.9550
501651278 × ⌀19.122.352.6715
80210168.38 × ⌀22.328.478.0840
1002552008 × ⌀22.331.7102.41127
150320269.912 × ⌀22.336.5154.21417
250445387.416 × ⌀28.447.4254.51863
Surface roughness (flange): Ra 3.2 to 6.3 µm
L
1)DN 8 with DN 15 flanges as standard
Endress+Hauser69
Page 70
Proline Promass F 300
Flange according to ASME B16.5: Class 300 with reduction in nominal diameter
1.4404 (F316/F316L)
DN
[mm]
5040AIS1651278 × ⌀ 19.122.352.6615
8050AKS210168.38 × ⌀ 22.328.478.0732
10080AMS2552008 × ⌀ 22.331.7102.4894
150100AOS320269.912 × ⌀ 22.336.5154.21187
200150AQS380330.212 × ⌀ 25.441.7202.71461
Surface roughness (flange): Ra 3.2 to 6.3 µm
Flange according to ASME B16.5: Class 600
1.4404 (F316/F316L): order code for "Process connection", option ACS
Alloy C22: order code for "Process connection", option ACC
Surface roughness (flange): Ra 3.2 to 6.3 µm
reduction
to DN
[mm]
DN
[mm]
1)
8
159566.74 × ⌀15.720.613.9420
2512588.94 × ⌀19.123.924.3490
40155114.34 × ⌀22.328.738.1600
501651278 × ⌀19.131.849.2742
80210168.38 × ⌀22.338.273.7900
100275215.98 × ⌀25.448.497.31157
150355292.112 × ⌀28.447.8154.21467
250510431.816 × ⌀35.169.9254.51946
[mm]
Order code for
"Process
connection",
option
A
9566.74 × ⌀15.720.613.9400
B
[mm]
A
[mm]B[mm]
C
[mm]
C
[mm]
D
[mm]
D
[mm]E[mm]L[mm]
E
[mm]
[mm]
L
1)DN 8 with DN 15 flanges as standard
Flange JIS B2220: 10K
1.4404 (F316/F316L): order code for "Process connection", option NDS
Alloy C22: order code for "Process connection", option NDC
DN
[mm]
501551204 × ⌀191650715
801851508 × ⌀191880832
1002101758 × ⌀19181001127
1502802408 × ⌀23221501354
25040035512 × ⌀25242501775
Surface roughness (flange): Ra 3.2 to 6.3 µm
A
[mm]
B
[mm]
C
[mm]
D
[mm]
E
[mm]
L
[mm]
70Endress+Hauser
Page 71
Proline Promass F 300
Flange JIS B2220: 20K
1.4404 (F316/F316L): order code for "Process connection", option NES
Alloy C22: order code for "Process connection", option NEC
DN
[mm]
1)
8
A
[mm]
B
[mm]
C
[mm]
D
[mm]
E
[mm]
95704 × ⌀151415370
1595704 × ⌀151415404
25125904 × ⌀191625440
401401054 × ⌀191840550
501551208 × ⌀191850715
802001608 × ⌀232280832
1002251858 × ⌀23241001127
15030526012 × ⌀25281501386
25043038012 × ⌀27342501845
Surface roughness (flange): Ra 1.6 to 3.2 µm
1)DN 8 with DN 15 flanges as standard
L
[mm]
Flange JIS B2220: 40K
1.4404 (F316/F316L): order code for "Process connection", option NGS
Alloy C22: order code for "Process connection", option NGC
DN
[mm]
1)
8
A
[mm]
B
[mm]
C
[mm]
D
[mm]
E
[mm]
115804 × ⌀192015400
15115804 × ⌀192015425
25130954 × ⌀192225485
401601204 × ⌀232438600
501651308 × ⌀192650760
802101708 × ⌀233275890
1002502058 × ⌀25361001167
15035529512 × ⌀33441501498
Surface roughness (flange): Ra 1.6 to 3.2 µm
1)DN 8 with DN 15 flanges as standard
Flange JIS B2220: 63K
1.4404 (F316/F316L): order code for "Process connection", option NHS
Alloy C22: order code for "Process connection", option NHC
DN
[mm]
1)
8
A
[mm]
B
[mm]
C
[mm]
D
[mm]
E
[mm]
120854 × ⌀192312420
15120854 × ⌀192312440
251401004 × ⌀232722494
401751304 × ⌀253235620
501851458 × ⌀233448775
802301858 × ⌀254073915
1002702208 × ⌀2744981167
L
[mm]
L
[mm]
Endress+Hauser71
Page 72
Proline Promass F 300
Flange JIS B2220: 63K
1.4404 (F316/F316L): order code for "Process connection", option NHS
Alloy C22: order code for "Process connection", option NHC
DN
[mm]
15036530512 × ⌀33541461528
Surface roughness (flange): Ra 1.6 to 3.2 µm
1)DN 8 with DN 15 flanges as standard
A
[mm]
B
[mm]
C
[mm]
D
[mm]
E
[mm]
L
[mm]
72Endress+Hauser
Page 73
Proline Promass F 300
C
D
A
B
E
XX
L
Fixed flange DIN 11864-2
A0015627
37Detail X: Asymmetrical process connection; the part shown in blue is provided by the supplier.
Length tolerance for dimension L in mm:
+1.5 / –2.0
Flange DIN11864-2 Form A, for pipe according to DIN11866 series A, flat with notch
1.4404 (316/316L)
Order code for "Process connection", option KCS
DN
[mm]
854374 × ⌀91010387
1559424 × ⌀91016418
2570534 × ⌀91026454
4082654 × ⌀91038560
5094774 × ⌀91050720
801331128 × ⌀111281900
1001591378 × ⌀11141001127
3A-version available: order code for "Additional approval", option LP in conjunction with
Ra ≤ 0.8 µm: order code for "Measuring tube material", option SB, SE or
Ra ≤ 0.4 µm: order code for "Measuring tube material", option SC, SF
A
[mm]
B
[mm]
C
[mm]
D
[mm]
E
[mm]
L
[mm]
Endress+Hauser73
Page 74
Proline Promass F 300
L
C
D
A
B
F
E
Lap joint flange EN 1092-1, ASME B16.5, JIS B2220
Length tolerance for dimension L in mm:
+1.5 / –2.0
Lap joint flange according to EN 1092-1 Form D: PN 40
1.4301 (F304), wetted parts Alloy C22Order code for "Process connection", option DAC
DN
[mm]A[mm]B[mm]
2)
8
95654 × ⌀ 1414.54517.33700
1595654 × ⌀ 1414.54517.34040
25115854 × ⌀ 1416.56828.5444+4
401501104 × ⌀ 18218843.1560+10
501651254 × ⌀ 182310254.5719+4
802001608 × ⌀ 182913882.5848+8
1002351908 × ⌀ 2234162107.11131+4
Surface roughness (flange): Ra 3.2 to 12.5 µm
C
[mm]
D
[mm]E[mm]F[mm]L[mm]
A0022221
L
diff
[mm]
1)
1)Difference to installation length of the welding neck flange (order code for "Process connection", option
D2C)
2)DN 8 with DN 15 flanges as standard
Lap joint flange according to ASME B16.5: Class 150
1.4301 (F304), wetted parts Alloy C22Order code for "Process connection", option ADC
DN
[mm]A[mm]B[mm]
2)
8
9060.34 × ⌀ 15.71535.115.73700
C
[mm]
D
[mm]E[mm]F[mm]L[mm]
L
diff
[mm]
1)
159060.34 × ⌀ 15.71535.115.74040
2511079.44 × ⌀ 15.71650.826.74400
4012598.44 × ⌀ 15.715.973.240.95500
50150120.74 × ⌀ 19.11991.952.67150
80190152.44 × ⌀ 19.122.3127.078.08400
74Endress+Hauser
Page 75
Proline Promass F 300
Lap joint flange according to ASME B16.5: Class 150
1.4301 (F304), wetted parts Alloy C22Order code for "Process connection", option ADC
DN
[mm]A[mm]
B
[mm]
C
[mm]
D
[mm]E[mm]F[mm]L[mm]
L
diff
[mm]
100230190.58 × ⌀ 19.126157.2102.411270
Surface roughness (flange): Ra 3.2 to 12.5 µm
1)Difference to installation length of the welding neck flange (order code for "Process connection", option
AAC)
2)DN 8 with DN 15 flanges as standard
Lap joint flange according to ASME B16.5: Class 300
1.4301 (F304), wetted parts Alloy C22Order code for "Process connection", option AEC
DN
[mm]A[mm]
2)
8
9566.74 × ⌀ 15.716.535.115.7376+6
B
[mm]
C
[mm]
D
[mm]E[mm]F[mm]L[mm]
L
diff
[mm]
159566.74 × ⌀ 15.716.535.115.7406+2
2512588.94 × ⌀ 19.121.050.826.7450+10
40155114.34 × ⌀ 22.323.073.240.9564+14
501651278 × ⌀ 19.125.591.952.6717+2
80210168.38 × ⌀ 22.331.0127.078.0852.6+12.6
1002552008 × ⌀ 22.332.0157.2102.41139+12
Surface roughness (flange): Ra 3.2 to 12.5 µm
1)
1)
1)Difference to installation length of the welding neck flange (order code for "Process connection", option
ABC)
2)DN 8 with DN 15 flanges as standard
Lap joint flange according to ASME B16.5: Class 600
1.4301 (F304), wetted parts Alloy C22Order code for "Process connection", option AFC
DN
[mm]A[mm]
2)
8
9566.74 × ⌀ 15.717.035.113.94000
B
[mm]
C
[mm]
D
[mm]E[mm]F[mm]L[mm]
L
diff
[mm]
159566.74 × ⌀ 15.717.035.113.94200
2512588.94 × ⌀ 19.121.550.824.34900
40155114.34 × ⌀ 22.325.073.238.16000
501651278 × ⌀ 19.128.091.949.27420
80210168.38 × ⌀ 22.335.0127.073.79000
100275215.98 × ⌀ 25.444.0157.297.31167+10
Surface roughness (flange): Ra 3.2 to 12.5 µm
1)Difference to installation length of the welding neck flange (order code for "Process connection", option
ACC)
2)DN 8 with DN 15 flanges as standard
1)
Endress+Hauser75
Page 76
Proline Promass F 300
Lap joint flange JIS B2220: 20K
1.4301 (F304), wetted parts Alloy C22Order code for "Process connection", option NIC
DN
[mm]
2)
8
1595704 × ⌀ 151451154040
25125904 × ⌀ 1918.567254400
401401054 × ⌀ 1918.581405500
501551208 × ⌀ 192396507150
802001608 × ⌀ 232913280844+12
1002251858 × ⌀ 232916010011270
Surface roughness (flange): Ra 3.2 to 12.5 µm
1)Difference to installation length of the welding neck flange (order code for "Process connection", option
NEC)
2)DN 8 with DN 15 flanges as standard
A
[mm]
B
[mm]
C
[mm]
D
[mm]
E
[mm]
F
[mm]
L
[mm]
95704 × ⌀ 151451153700
L
diff
[mm]
1)
76Endress+Hauser
Page 77
Proline Promass F 300
L
A
B
Clamp connections
Tri-Clamp
Length tolerance for dimension L in mm:
+1.5 / –2.0
Tri-Clamp (½"), for pipe according to DIN 11866 series C
1.4404 (316/316L)
Order code for "Process connection", option FDW
DN
[mm]
8½25.09.5367
15½25.09.5398
3-A version available: order code for "Additional approval", option LP in conjunction with
Ra ≤ 0.8 µm: order code for "Measuring tube material", option SB, SE or
Ra ≤ 0.4 µm: order code for "Measuring tube material", option SC, SF
Clamp
[in]
A
[mm]
B
[mm]
A0015625
L
[mm]
Tri-Clamp ( ≥ 1"), for pipe according to DIN 11866 series C
1.4404 (316/316L)
Order code for "Process connection", option FTS
DN
[mm]
8150.422.1367
15150.422.1398
25150.422.1434
401½50.434.8560
50263.947.5720
80390.972.9900
1004118.997.41127
3-A version available: order code for "Additional approval", option LP in conjunction with
Ra ≤ 0.8 µm: order code for "Measuring tube material", option SB, SE or
Ra ≤ 0.4 µm: order code for "Measuring tube material", option SC, SF
Clamp
[in]
A
[mm]
B
[mm]
L
[mm]
Endress+Hauser77
Page 78
Threaded couplings
L
A
B
Thread DIN 11851, DIN11864-1, SMS 1145
Length tolerance for dimension L in mm:
+1.5 / –2.0
Thread DIN 11851, for pipe according to DIN11866, series A
1.4404 (316/316L)
Order code for "Process connection", option FMW
DN
[mm]
8Rd 34 × ¹⁄₈16367
15Rd 34 × ¹⁄₈16398
25Rd 52 × ¹⁄₆26434
40Rd 65 × ¹⁄₆38560
50Rd 78 × ¹⁄₆50720
80Rd 110 × ¹⁄₄81900
100Rd 130 × ¹⁄₄1001127
3-A version available: order code for "Additional approval", option LP in conjunction with
Ra ≤ 0.8 µm: order code for "Measuring tube material", option SB, SE
A
[in]
B
[mm]
Proline Promass F 300
A0015628
L
[mm]
Thread DIN11864-1 Form A, for pipe according to DIN11866, series A
1.4404 (316/316L)
Order code for "Process connection", option FLW
DN
[mm]
8Rd 28 × ¹⁄₈10367
15Rd 34 × ¹⁄₈16398
25Rd 52 × ¹⁄₈26434
40Rd 65 × ¹⁄₆38560
50Rd 78 × ¹⁄₆50720
80Rd 110 × ¹⁄₄81900
100Rd 130 × ¹⁄₄1001127
3-A version available: order code for "Additional approval", option LP in conjunction with
Ra ≤ 0.8 µm: order code for "Measuring tube material", option SB, SE or
Ra ≤ 0.4 µm: order code for "Measuring tube material", option SC, SF
A
[in]
B
[mm]
L
[mm]
78Endress+Hauser
Page 79
Proline Promass F 300
Thread SMS 1145
1.4404 (316/316L)
Order code for "Process connection", option SCS
DN
[mm]
8Rd 40 × ¹⁄₆22.6367
15Rd 40 × ¹⁄₆22.6398
25Rd 40 × ¹⁄₆22.6434
40Rd 60 × ¹⁄₆35.6560
50Rd 70 × ¹⁄₆48.6720
80Rd 98 × ¹⁄₆72.9900
100Rd 132 × ¹⁄₆97.61127
3-A version available: order code for "Additional approval", option LP in conjunction with
Ra ≤ 0.8 µm: order code for "Measuring tube material", option SB, SE
A
[in]
B
[mm]
L
[mm]
Endress+Hauser79
Page 80
Thread ISO 2853
L
A
B
Length tolerance for dimension L in mm:
+1.5 / –2.0
Thread ISO 2853, for pipe according to ISO 2037
1.4404 (316/316L)
Order code for "Process connection", option JSF
DN
[mm]
837.1322.6367
1537.1322.6398
2537.1322.6434
4052.6835.6560
5064.1648.6720
8091.1972.9900
100118.2197.61127
3-A version available: order code for "Additional approval", option LP in conjunction with
Ra ≤ 0.8 µm: order code for "Measuring tube material", option SB, SE or
Ra ≤ 0.4 µm: order code for "Measuring tube material", option SC, SF
1)
A
[mm]
B
[mm]
Proline Promass F 300
A0015623
L
[mm]
1)Max. thread diameter as per ISO 2853 annex A
80Endress+Hauser
Page 81
Proline Promass F 300
L
A
B
VCO
Length tolerance for dimension L in mm:
+1.5 / –2.0
8-VCO-4 (½")
1.4404 (316/316L)
Order code for "Process connection", option CVS
DN
[mm]
8AF 110.2390
A
[in]
B
[mm]
A0015624
L
[mm]
12-VCO-4 (¾")
1.4404 (316/316L)
Order code for "Process connection", option CWS
DN
[mm]
15AF 1½15.7430
A
[in]
B
[mm]
L
[mm]
Endress+Hauser81
Page 82
Accessories
DN 8 ( ")...150 (6")!⁄#DN 250 (10")
LL
45°
A
A
SW 1"
2
½" NPT
1
½" NPT
35 (1.38)ca. 75 (approx. 2.95)
1
1;2
1
1;2
146 (5.75)
48 (1.9)
12 (0.47)
280 (11.0)255 (10.0)
134 (5.3)
30 (1.18)
Rupture disk/purge connections
Proline Promass F 300
38
1Connection nipple for purge connections: order code for "Sensor options", option CH "Purge connection"
2Connection nipple with rupture disk: order code for "Sensor option", option CA "Rupture disk"
DNAL
[mm][mm][mm]
862216
1562220
2562260
4067310
5079452
80101560
100120684
150141880
250182380
Protective cover
A0028914
82Endress+Hauser
A0029553
Page 83
Proline Promass F 300
136 (5.35)
114 (4.49)
86 (3.39)
78 (3.07)
!103 (4.06)
!12 (0.47)
7.0 (0.27)
68 (2.7)105 (4.1)
173 (6.8)
Remote display and operating module DKX001
A0028921
39Engineering unit mm (in)
External WLAN antenna
The external WLAN antenna is not suitable for use in hygienic applications.
External WLAN antenna mounted on device
A0028923
40Engineering unit mm (in)
External WLAN antenna mounted with cable
The external WLAN antenna can be mounted separately from the transmitter if the transmission/
reception conditions at the transmitter mounting location are poor.
Endress+Hauser83
Page 84
70 (2.8)
1500 (59.1)
72 (2.8)
41Engineering unit mm (in)
Proline Promass F 300
A0033597
84Endress+Hauser
Page 85
Proline Promass F 300
D
L
E
F
A
CB
K
G
H
I
M
Dimensions in US unitsCompact version
Order code for "Housing", option A "Aluminum, coated"
1)Depending on the cable gland used: values up to + 1.18 in
2)With order code for "Sensor option", option CG or order code for "Measuring tube material", option SD, SE,
SF, TH, LA: values +2.76 in
3)With order code for "Measuring tube material", option TT, TU: values +4.09 in
4)Depends on the process connection in question
2) 3)
FGHIKLM
4)
2.76
4)
2.76
4)
2.76
4)
3.11
4)
3.90
4)
5.47
4)
6.93
4)
8.58
4)
12.01
86Endress+Hauser
Page 87
Proline Promass F 300
C
D
A
B
E
L
Flange connections
Fixed flange ASME B16.5
A0015621
Length tolerance for dimension L in inch:
• DN ≤ 4": +0.06 / –0.08
• DN ≥ 5": +0.14
Flange according to ASME B16.5: Class 150
1.4404 (F316/F316L): order code for "Process connection", option AAS
Alloy C22: order code for "Process connection", option AAC
DN
[in]
1)
³⁄₈
½3.542.374 × ⌀0.620.440.6215.91
14.333.134 × ⌀0.620.561.0517.32
1½4.923.874 × ⌀0.620.691.6121.65
25.914.754 × ⌀0.750.752.0728.15
37.486.004 × ⌀0.750.943.0733.07
49.067.508 × ⌀0.750.944.0344.37
611.029.508 × ⌀0.8816.0755.04
1015.9414.2512 × ⌀1.01.1910.0272.13
Surface roughness (flange): Ra 125 to 250 µin
1)DN ³⁄₈" with DN ½" flanges as standard
Flange according to ASME B16.5: Class 150 with reduction in nominal diameter
1.4404 (F316/F316L)
DN
[in]
21½AHS5.914.754 × ⌀ 0.750.752.0721.65
32AJS7.4864 × ⌀ 0.750.943.0728.35
43ALS9.067.58 × ⌀ 0.750.944.0334.41
64ANS11.029.58 × ⌀ 0.8816.0745.94
86APS13.5811.758 × ⌀ 0.881.147.9857.52
Surface roughness (flange): Ra 125 to 250 µin
A
[in]
3.542.374 × ⌀ 0.620.440.6214.57
reduction
to DN
[in]
Order code for
B
[in]
"Process
connection",
option
A
[in]
C
[in]
B
[in]
D
[in]
C
[in]
E
[in]
D
[in]
E
[in]
L
[in]
L
[in]
Endress+Hauser87
Page 88
Proline Promass F 300
Flange according to ASME B16.5: Class 300
1.4404 (F316/F316L): order code for "Process connection", option ABS
Alloy C22: order code for "Process connection", option ABC
DN
[in]
1)
³⁄₈
½3.742.634 × ⌀0.620.560.6215.91
14.923.504 × ⌀0.750.691.0517.32
1½6.104.504 × ⌀0.880.811.6121.65
26.505.008 × ⌀0.750.882.0728.15
38.276.638 × ⌀0.881.123.0733.07
410.047.878 × ⌀0.881.254.0344.37
612.610.6312 × ⌀0.881.446.0755.79
1017.5215.2516 × ⌀1.121.8710.0273.35
Surface roughness (flange): Ra 125 to 250 µin
1)DN ³⁄₈" with DN ½" flanges as standard
A
[in]
B
[in]
C
[in]
D
[in]
E
[in]
3.742.634 × ⌀0.620.560.6214.57
L
[in]
Flange according to ASME B16.5: Class 300 with reduction in nominal diameter
1.4404 (F316/F316L)
DN
[in]
reduction
to DN
[in]
Order code for
"Process
connection",
A
[in]
B
[in]
C
[in]
D
[in]
E
[in]
option
21½AIS6.558 × ⌀ 0.750.882.0724.21
32AKS8.276.638 × ⌀ 0.881.123.0728.82
43AMS10.047.878 × ⌀ 0.881.254.0335.2
64AOS12.610.6312 × ⌀ 0.881.446.0746.73
86AQS14.961312 × ⌀ 11.647.9857.52
Surface roughness (flange): Ra 125 to 250 µin
Flange according to ASME B16.5: Class 600
1.4404 (F316/F316L): order code for "Process connection", option ACS
Alloy C22: order code for "Process connection", option ACC
DN
[in]
³⁄₈
1)
A
[in]
B
[in]
C
[in]
D
[in]
E
[in]
3.742.634 × ⌀0.620.810.5515.75
½3.742.634 × ⌀0.620.810.5516.54
14.923.504 × ⌀0.750.940.9619.29
1½6.104.504 × ⌀0.881.131.523.62
26.505.008 × ⌀0.751.251.9429.21
38.276.638 × ⌀0.881.52.935.43
410.838.508 × ⌀1.001.913.8345.55
613.9811.5012 × ⌀1.121.886.0757.76
1020.0817.0016 × ⌀1.382.7510.0276.61
Surface roughness (flange): Ra 125 to 250 µin
L
[in]
L
[in]
1)DN ³⁄₈" with DN ½" flanges as standard
88Endress+Hauser
Page 89
Proline Promass F 300
L
C
D
A
B
F
E
Lap joint flange ASME B16.5
Length tolerance for dimension L in inch:
+0.06 / –0.08
Lap joint flange according to ASME B16.5: Class 150
1.4301 (F304), wetted parts Alloy C22Order code for "Process connection", option ADC
DN
[in]
2)
³⁄₈
½3.542.374 × ⌀ 0.620.591.380.6215.910
14.333.134 × ⌀ 0.620.6321.0517.320
1½4.923.874 × ⌀ 0.620.632.881.6121.650
25.914.754 × ⌀ 0.750.753.622.0728.150
37.486.004 × ⌀ 0.750.8853.0733.070
49.067.508 × ⌀ 0.751.026.194.0344.370
Surface roughness (flange): Ra 125 to 492 µin
A
[in]
B
[in]
C
[in]
D
[in]
E
[in]
F
[in]
L
[in]
3.542.374 × ⌀ 0.620.591.380.6214.570
L
diff
[in]
A0022221
1)
1)Difference to installation length of the welding neck flange (order code for "Process connection", option
AAC)
2)DN ³⁄₈" with DN ½" flanges as standard
Lap joint flange according to ASME B16.5: Class 300
1.4301 (F304), wetted parts Alloy C22Order code for "Process connection", option AEC
DN
[in]
³⁄₈
A
[in]
2)
3.742.634 × ⌀ 0.620.651.380.6214.8+0.23
B
[in]
C
[in]
D
[in]
E
[in]
F
[in]
L
[in]
L
diff
[in]
1)
½3.742.634 × ⌀ 0.620.651.380.6215.98+0.07
14.923.504 × ⌀ 0.750.8321.0517.72+0.40
1½6.104.504 × ⌀ 0.880.912.881.6122.2+0.55
26.505.008 × ⌀ 0.7513.622.0728.23+0.08
38.276.638 × ⌀ 0.881.2253.0733.57+0.50
Endress+Hauser89
Page 90
Proline Promass F 300
Lap joint flange according to ASME B16.5: Class 300
1.4301 (F304), wetted parts Alloy C22Order code for "Process connection", option AEC
DN
[in]
410.047.878 × ⌀ 0.881.266.194.0344.84+0.47
Surface roughness (flange): Ra 125 to 492 µin
1)Difference to installation length of the welding neck flange (order code for "Process connection", option
AAC)
2)DN ³⁄₈" with DN ½" flanges as standard
Lap joint flange according to ASME B16.5, Class 600
1.4301 (F304), wetted parts Alloy C22Order code for "Process connection", option AFC
DN
[in]
2)
³⁄₈
½3.742.634 × ⌀ 0.620.671.380.5516.540
14.923.504 × ⌀ 0.750.8520.9619.290
1½6.104.504 × ⌀ 0.880.982.881.523.620
26.505.008 × ⌀ 0.751.13.621.9429.210
38.276.638 × ⌀ 0.881.3852.935.430
410.838.508 × ⌀ 11.736.193.8345.94+0.39
Surface roughness (flange): Ra 125 to 492 µin
A
[in]
A
[in]
B
[in]
B
[in]
C
[in]
C
[in]
D
[in]
D
[in]
E
[in]
E
[in]
F
[in]
F
[in]
L
[in]
L
[in]
L
[in]
L
[in]
3.742.634 × ⌀ 0.620.671.380.5515.750
diff
diff
1)
1)
1)Difference to installation length of the welding neck flange (order code for "Process connection", option
AAC)
2)DN ³⁄₈" with DN ½" flanges as standard
90Endress+Hauser
Page 91
Proline Promass F 300
L
A
B
Clamp connections
Tri-Clamp
Length tolerance for dimension L in inch:
+0.06 / –0.08
Tri-Clamp (½"), DIN 11866 series C
1.4404 (316/316L)
Order code for "Process connection", option FDW
DN
[in]
³⁄₈½0.980.3714.4
½½0.980.3715.7
3-A version available: order code for "Additional approval", option LP in conjunction with
Ra ≤ 32 µin: order code for "Measuring tube material", option SB, SE or
Ra ≤ 16 µin: order code for "Measuring tube material", option SC, SF
Clamp
[in]
A
[in]
B
[in]
A0015625
L
[in]
Tri-Clamp ( ≥ 1"), DIN 11866 series C
1.4404 (316/316L)
Order code for "Process connection", option FTS
DN
[in]
³⁄₈11.980.8714.4
½11.980.8715.7
111.980.8717.1
1½1½1.981.3722.0
222.521.8728.3
333.582.8735.4
444.683.8344.4
3-A version available: order code for "Additional approval", option LP in conjunction with
Ra ≤ 32 µin: order code for "Measuring tube material", option SB, SE or
Ra ≤ 16 µin: order code for "Measuring tube material", option SC, SF
Clamp
[in]
A
[in]
B
[in]
L
[in]
Endress+Hauser91
Page 92
Threaded couplings
L
A
B
Thread SMS 1145
Length tolerance for dimension L in inch:
+0.06 / –0.08
Thread SMS 1145
1.4404 (316/316L)
Order code for "Process connection", option SCS
DN
[in]
³⁄₈Rd 40 × ¹⁄₆0.8914.45
½Rd 40 × ¹⁄₆0.8915.67
1Rd 40 × ¹⁄₆0.8917.09
1½Rd 60 × ¹⁄₆1.422.05
2Rd 70 × ¹⁄₆1.9128.35
3Rd 98 × ¹⁄₆2.8735.43
4Rd 132 × ¹⁄₆3.8444.37
3-A version available: order code for "Additional approval", option LP in conjunction with
Ra ≤ 32 µin: order code for "Measuring tube material", option SB, SE
A
[in]
B
[in]
Proline Promass F 300
A0015628
L
[in]
92Endress+Hauser
Page 93
Proline Promass F 300
L
A
B
DN 8 ( ")...150 (6")!⁄#DN 250 (10")
LL
45°
A
A
SW 1"
2
½" NPT
1
½" NPT
35 (1.38)ca. 75 (approx. 2.95)
1
1;2
1
1;2
VCO
Length tolerance for dimension L in inch:
+0.06 / –0.08
8-VCO-4 (½")
1.4404 (316/316L)
Order code for "Process connection", option CVS
DN
[in]
³⁄₈AF 10.415.35
A
[in]
B
[in]
A0015624
L
[in]
12-VCO-4 (¾")
1.4404 (316/316L)
Order code for "Process connection", option CWS
Accessories
Rupture disk/purge connections
DN
[in]
½AF 1½0.6216.93
A
[in]
B
[in]
L
[in]
Endress+Hauser93
1Connection nipple for purge connections: order code for "Sensor options", option CH "Purge connection"
2Connection nipple with rupture disk: order code for "Sensor option", option CA "Rupture disk"
A0028914
Page 94
Protective cover
146 (5.75)
48 (1.9)
12 (0.47)
280 (11.0)255 (10.0)
134 (5.3)
30 (1.18)
136 (5.35)
114 (4.49)
86 (3.39)
78 (3.07)
!103 (4.06)
!12 (0.47)
7.0 (0.27)
Proline Promass F 300
DNAL
[in][in][in]
³⁄₈2.448.50
½2.448.66
12.4410.24
1½2.6412.20
23.1117.78
33.9822.0
44.7227.0
65.5534.6
107.1714.96
Remote display and operating module DKX001
42Engineering unit mm (in)
External WLAN antenna
The external WLAN antenna is not suitable for use in hygienic applications.
A0029553
A0028921
94Endress+Hauser
Page 95
Proline Promass F 300
68 (2.7)105 (4.1)
173 (6.8)
70 (2.8)
1500 (59.1)
72 (2.8)
External WLAN antenna mounted on device
A0028923
43Engineering unit mm (in)
External WLAN antenna mounted with cable
The external WLAN antenna can be mounted separately from the transmitter if the transmission/
reception conditions at the transmitter mounting location are poor.
A0033597
44Engineering unit mm (in)
Weight
All values (weight exclusive of packaging material) refer to devices with EN/DIN PN 40 flanges.
Weight specifications including transmitter as per order code for "Housing", option A "Aluminum,
coated".
Different values due to different transmitter versions:
• Transmitter version for the hazardous area
(Order code for "Housing", option A "Aluminum, coated"; Ex d): +2 kg (+4.4 lbs)
• Cast transmitter version, stainless
(Order code for "Housing", option L "Cast, stainless"): +6 kg (+13 lbs)
• Transmitter version for hygienic area
(Order code for "Housing", option B "Stainless, hygienic"): +0.2 kg (+0.44 lbs)
Weight in SI units
DN [mm]Weight [kg]
811
1512
2514
4019
Endress+Hauser95
5030
Page 96
DN [mm]Weight [kg]
Weight in US units
DN [in]Weight [lbs]
Proline Promass F 300
8055
10096
150154
250400
3/824
½26
131
1½42
266
3121
4212
6340
10882
MaterialsTransmitter housing
Order code for "Housing":
• Option A "Aluminum, coated": aluminum, AlSi10Mg, coated
• Option B "Stainless, hygienic": stainless steel, 1.4404 (316L)
• Option L "Cast, stainless": cast, stainless steel, 1.4409 (CF3M) similar to 316L
Window material
Order code for "Housing":
• Option A "Aluminum, coated": glass
• Option B "Stainless, hygienic": polycarbonate
• Option L "Cast, stainless": glass
Seals
Order code for "Housing":
Option B "Stainless, hygienic": EPDM and silicone
96Endress+Hauser
Page 97
Proline Promass F 300
1
2
3
4
Cable entries/cable glands
A0028352
45Possible cable entries/cable glands
1Female thread M20 × 1.5
2Cable gland M20 × 1.5
3Adapter for cable entry with internal thread G ½" or NPT ½"
4Device plugs
Order code for "Housing", option A "Aluminum, coated"
The various cable entries are suitable for hazardous and non-hazardous areas.
Cable entry/cable glandMaterial
Cable gland M20 × 1.5Plastic/nickel-plated brass
Adapter for cable entry with internal thread G ½"Nickel-plated brass
Adapter for cable entry with internal thread NPT ½"
Device plugPlug M12 × 1
• Socket: Stainless steel, 1.4404 (316L)
• Contact housing: Polyamide
• Contacts: Gold-plated brass
Order code for "Housing", option B "Stainless, hygienic"
The various cable entries are suitable for hazardous and non-hazardous areas.
Cable entry/cable glandMaterial
Cable gland M20 × 1.5Plastic
Adapter for cable entry with internal thread G ½"Nickel-plated brass
Adapter for cable entry with internal thread NPT ½"
Device plugPlug M12 × 1
• Socket: Stainless steel, 1.4404 (316L)
• Contact housing: Polyamide
• Contacts: Gold-plated brass
Order code for "Housing", option L "Cast, stainless"
The various cable entries are suitable for hazardous and non-hazardous areas.
• Flanges according to EN 1092-1 (DIN2501) / according to ASME B 16.5 / as per JIS B2220:
– Stainless steel, 1.4404 (F316/F316L)
– Alloy C22, 2.4602 (UNS N06022)
– Lap joint flanges: stainless steel, 1.4301 (F304); wetted parts Alloy C22
• All other process connections:
Stainless steel, 1.4404 (316/316L)
High-temperature version
Flanges according to EN 1092-1 (DIN2501) / according to ASME B 16.5 / as per JIS B2220:
– Stainless steel, 1.4404 (F316/F316L)
– Alloy C22, 2.4602 (UNS N06022)
Available process connections→ 99
Seals
Welded process connections without internal seals
98Endress+Hauser
Page 99
Proline Promass F 300
Accessories
Protective cover
Stainless steel, 1.4404 (316L)
External WLAN antenna
• Antenna: ASA plastic (acrylic ester-styrene-acrylonitrile) and nickel-plated brass
• Adapter: Stainless steel and nickel-plated brass
• Cable: Polyethylene
• Plug: Nickel-plated brass
• Angle bracket: Stainless steel
Process connections
Surface roughness
• Fixed flange connections:
– EN 1092-1 (DIN 2501) flange
– EN 1092-1 (DIN 2512N) flange
– Namur lengths in accordance with NE 132
– ASME B16.5 flange
– JIS B2220 flange
– DIN 11864-2 Form A flange, DIN 11866 series A, flange with notch
• Clamp connections:
Tri-Clamp (OD tubes), DIN 11866 series C
• Thread:
– DIN 11851 thread, DIN 11866 series A
– SMS 1145 thread
– ISO 2853 thread, ISO 2037
– DIN 11864-1 Form A thread, DIN 11866 series A
• VCO connections:
– 8-VCO-4
– 12-VCO-4
Process connection materials → 98
All data relate to parts in contact with fluid. The following surface roughness quality can be ordered.
• Not polished
• Ra
= 0.8 µm (32 µin)
max
• Ra
= 0.4 µm (16 µin)
max
Operability
Operating concept
Endress+Hauser99
Operator-oriented menu structure for user-specific tasks
• Commissioning
• Operation
• Diagnostics
• Expert level
Fast and safe commissioning
• Guided menus ("Make-it-run" wizards) for applications
• Menu guidance with brief descriptions of the individual parameter functions
• Device access via Web server or SmartBlue app → 119
• WLAN access to the device via mobile handheld terminal, tablet or smart phone
Reliable operation
• Operation in local language → 100
• Uniform operating philosophy applied to device and operating tools
• If replacing electronic modules, transfer the device configuration via the integrated memory
(HistoROM backup) which contains the process and measuring device data and the event logbook.
No need to reconfigure.
• Troubleshooting measures can be called up via the device and in the operating tools
• Diverse simulation options, logbook for events that occur and optional line recorder functions
Page 100
Proline Promass F 300
Languages
Can be operated in the following languages:
• Via local operation
English, German, French, Spanish, Italian, Dutch, Portuguese, Polish, Russian, Turkish, Chinese,
Japanese, Korean, Bahasa (Indonesian), Vietnamese, Czech, Swedish
• Via Web browser
English, German, French, Spanish, Italian, Dutch, Portuguese, Polish, Russian, Turkish, Chinese,
Japanese, Korean, Bahasa (Indonesian), Vietnamese, Czech, Swedish
• Via "FieldCare", "DeviceCare" operating tool: English, German, French, Spanish, Italian, Chinese,
Japanese
Local operationVia display module
Two display modules are available:
• Order code for "Display; operation", option F "4-line, illuminated, graphic display; touch control"
• Order code for "Display; operation", option G "4-line, illuminated, graphic display; touch control +
WLAN"
Information about WLAN interface → 107
46Operation with touch control
Display elements
• 4-line, illuminated, graphic display
• White background lighting; switches to red in event of device errors
• Format for displaying measured variables and status variables can be individually configured
• Permitted ambient temperature for the display: –20 to +60 °C (–4 to +140 °F)
The readability of the display may be impaired at temperatures outside the temperature range.
Operating elements
• External operation via touch control (3 optical keys) without opening the housing: , ,
• Operating elements also accessible in the various zones of the hazardous area
Via remote display and operating module DKX001
The remote display and operating module DKX001 is available as an optional extra → 117.
• The remote display and operating module DKX001 is only available for the following housing
versions, order code for "Housing":
– Option A "Aluminum, coated"
– Option L "Cast, stainless"
• The measuring device is always supplied with a dummy cover when the remote display and
operating module DKX001 is ordered directly with the measuring device. Display or operation
at the transmitter is not possible in this case.
• If ordered subsequently, the remote display and operating module DKX001 may not be
connected at the same time as the existing measuring device display module. Only one
display or operation unit may be connected to the transmitter at any one time.
A0026785
100Endress+Hauser
Loading...
+ hidden pages
You need points to download manuals.
1 point = 1 manual.
You can buy points or you can get point for every manual you upload.