Endress+Hauser F 300 Specifications

Page 1
TI01221D/06/EN/05.18 71402583 2018-06-01
Products Solutions Services
Technical Information
Proline Promass F 300
Coriolis flowmeter
The flowmeter with premium accuracy, robustness and a compact, easily accessible transmitter
Application
• Measuring principle operates independently of physical fluid properties such as viscosity or density
• Highest measurement performance for liquids and gases under varying, demanding process conditions
Device properties
• Medium temperature: –196 to +350 °C (–320 to +662 °F)
• Nominal diameter: DN 8 to 250 (³⁄₈ to 10")
• Compact dualcompartment housing with up to 3 I/Os
• Backlit display with touch control and WLAN access
• Remote display available
Your benefits
• Highest process safety – immune to fluctuating and harsh environments
• Fewer process measuring points – multivariable measurement (flow, density, temperature)
• Spacesaving installation – no inlet/outlet run needs
• Full access to process and diagnostic information – numerous, freely combinable I/Os and fieldbuses
• Reduced complexity and variety – freely configurable I/O functionality
• Integrated verification – Heartbeat Technology
Page 2

Table of contents

Proline Promass F 300
About this document ........................ 4
Symbols used ................................ 4
Function and system design ................... 5
Measuring principle ............................ 5
Measuring system ............................. 7
Equipment architecture ......................... 8
Safety ..................................... 8
Input .................................... 11
Measured variable ............................ 11
Measuring range ............................. 11
Operable flow range ........................... 12
Input signal ................................ 12
Output .................................. 14
Output and input variants ....................... 14
Output signal ............................... 15
Signal on alarm .............................. 19
Ex connection data ........................... 21
Low flow cut off ............................. 22
Galvanic isolation ............................ 22
Protocol-specific data .......................... 22
Power supply ............................. 27
Terminal assignment .......................... 27
Device plugs available .......................... 28
Pin assignment, device plug ...................... 29
Supply voltage .............................. 30
Power consumption ........................... 31
Current consumption .......................... 31
Power supply failure .......................... 31
Electrical connection .......................... 32
Potential equalization ......................... 41
terminals .................................. 41
Cable entries ............................... 41
Cable specification ............................ 41
Performance characteristics .................. 44
Reference operating conditions ................... 44
Maximum measured error ....................... 44
Repeatability ............................... 46
Response time .............................. 46
Influence of ambient temperature ................. 46
Influence of medium temperature .................. 47
Influence of medium pressure .................... 47
Design fundamentals .......................... 48
Installation ............................... 48
Mounting location ............................ 49
Orientation ................................ 50
Inlet and outlet runs .......................... 50
Special mounting instructions .................... 50
Environment .............................. 52
Ambient temperature range ..................... 52
Storage temperature .......................... 52
Climate class ............................... 52
Degree of protection .......................... 52
Vibration resistance ........................... 52
Shock resistance ............................. 52
Shock resistance ............................. 52
Interior cleaning ............................. 52
Electromagnetic compatibility (EMC) ............... 52
Process .................................. 53
Medium temperature range ...................... 53
Density ................................... 54
Pressure-temperature curves ..................... 54
Sensor housing .............................. 59
Rupture disk ................................ 60
Flow limit ................................. 60
Pressure loss ............................... 60
System pressure ............................. 60
Thermal insulation ........................... 61
Heating ................................... 61
Vibrations ................................. 62
Custody transfer measurement ............... 63
Mechanical construction .................... 64
Dimensions in SI units ......................... 64
Dimensions in US units ......................... 85
Weight ................................... 95
Materials .................................. 96
Process connections ........................... 99
Surface roughness ........................... 99
Operability ............................... 99
Operating concept ............................ 99
Languages ................................ 100
Local operation ............................. 100
Remote operation ........................... 101
Service interface ............................ 107
Network integration .......................... 108
Supported operating tools ...................... 109
HistoROM data management .................... 110
Certificates and approvals .................. 111
CE mark .................................. 111
C-Tick symbol .............................. 111
Ex approval ............................... 111
Sanitary compatibility ......................... 113
Pharmaceutical compatibility .................... 113
Functional safety ............................ 113
HART certification ........................... 113
FOUNDATION Fieldbus certification ............... 113
Certification PROFIBUS ........................ 113
EtherNet/IP certification ....................... 113
Certification PROFINET ........................ 113
Pressure Equipment Directive ................... 114
Radio approval ............................. 114
Measuring instrument approval .................. 114
Additional certification ........................ 114
Other standards and guidelines .................. 115
2 Endress+Hauser
Page 3
Proline Promass F 300
Ordering information ...................... 115
Application packages ...................... 116
Diagnostics functions ......................... 116
Heartbeat Technology ........................ 116
Concentration .............................. 116
Special density ............................. 116
Petroleum ................................ 117
OPC-UA server ............................. 117
Accessories .............................. 117
Device-specific accessories ...................... 117
Communication-specific accessories ............... 118
Service-specific accessories ..................... 119
System components .......................... 119
Supplementary documentation .............. 120
Standard documentation ....................... 120
Device-dependent additional documentation ......... 121
Registered trademarks ..................... 122
Endress+Hauser 3
Page 4

About this document

A

Symbols used Electrical symbols

Symbol Meaning
Proline Promass F 300
Direct current
Alternating current
Direct current and alternating current
Ground connection
A grounded terminal which, as far as the operator is concerned, is grounded via a grounding system.
Protective Earth (PE)
A terminal which must be connected to ground prior to establishing any other connections.
The ground terminals are situated inside and outside the device:
• Inner ground terminal: Connects the protectiv earth to the mains supply.
• Outer ground terminal: Connects the device to the plant grounding system.
Communication symbols
Symbol Meaning
Wireless Local Area Network (WLAN)
Communication via a wireless, local network.
LED
Light emitting diode is off.
LED
Light emitting diode is on.
LED
Light emitting diode is flashing.
Symbols for certain types of information
Symbol Meaning
Permitted
Procedures, processes or actions that are permitted.
Preferred
Procedures, processes or actions that are preferred.
Forbidden
Procedures, processes or actions that are forbidden.
Tip
Indicates additional information.
Reference to documentation.
Reference to page.
Reference to graphic.
Visual inspection.
4 Endress+Hauser
Page 5
Proline Promass F 300
1.
-
.
21 3
Symbols in graphics
Symbol Meaning
1, 2, 3, ... Item numbers
, 2., 3., … Series of steps
A, B, C, ... Views
A-A, B-B, C-C, ... Sections
Hazardous area
Safe area (non-hazardous area)
Flow direction

Function and system design

Measuring principle

The measuring principle is based on the controlled generation of Coriolis forces. These forces are always present in a system when both translational and rotational movements are superimposed.
Fc = 2 · ∆m (ν · ω)
Fc = Coriolis force
∆m = moving mass
ω = rotational velocity
ν = radial velocity in rotating or oscillating system
The amplitude of the Coriolis force depends on the moving mass ∆m, its velocity ν in the system and thus on the mass flow. Instead of a constant rotational velocity ω, the sensor uses oscillation.
In the sensor, two parallel measuring tubes containing flowing fluid oscillate in antiphase, acting like a tuning fork. The Coriolis forces produced at the measuring tubes cause a phase shift in the tube oscillations (see illustration):
• At zero flow (when the fluid is at a standstill) the two tubes oscillate in phase (1).
• Mass flow causes deceleration of the oscillation at the inlet of the tubes (2) and acceleration at the outlet (3).
A0028850
The phase difference (A-B) increases with increasing mass flow. Electrodynamic sensors register the tube oscillations at the inlet and outlet. System balance is ensured by the antiphase oscillation of the two measuring tubes. The measuring principle operates independently of temperature, pressure, viscosity, conductivity and flow profile.
Endress+Hauser 5
Page 6
Proline Promass F 300
Density measurement
The measuring tube is continuously excited at its resonance frequency. A change in the mass and thus the density of the oscillating system (comprising measuring tube and fluid) results in a corresponding, automatic adjustment in the oscillation frequency. Resonance frequency is thus a function of medium density. The microprocessor utilizes this relationship to obtain a density signal.
Volume measurement
Together with the measured mass flow, this is used to calculate the volume flow.
Temperature measurement
The temperature of the measuring tube is determined in order to calculate the compensation factor due to temperature effects. This signal corresponds to the process temperature and is also available as an output signal.
6 Endress+Hauser
Page 7
Proline Promass F 300

Measuring system

The device consists of a transmitter and a sensor.
The device is available as a compact version: The transmitter and sensor form a mechanical unit.
Transmitter
Promass 300
Device versions and materials:
• Transmitter housing – Aluminum, coated: aluminum, AlSi10Mg, coated – Stainless, hygienic: stainless steel, 1.4404 – Cast, stainless: cast, stainless steel, 1.4409 (CF3M) similar to 316L
• Material of window in transmitter housing: – Aluminum, coated: glass
A0026708
– Stainless, hygienic: polycarbonate – Cast, stainless: glass
Configuration:
• External operation via 4-line, backlit, graphic local display with touch control and guided menus ("Make-it-run" wizards) for application­specific commissioning.
• Via service interface or WLAN interface: – Operating tools (e.g. FieldCare, DeviceCare, SmartBlue app) – Web server (access via Web browser, e.g. Microsoft Internet
Explorer, Microsoft Edge)
Sensor
PromassF • Excellent performance across a wide range of applications
• Simultaneous measurement of flow, volume flow, density and temperature (multivariable)
• Immune to process influences
• Nominal diameter range: DN 8 to 250 (³⁄₈ to 10")
• Materials:
A0026713
– Sensor: stainless steel, 1.4301/1.4307 (304L); optional 1.4404
(316/316L)
– Measuring tubes: stainless steel, 1.4539 (904L); 1.4404
(316/316L); Alloy C22, 2.4602 (UNS N06022)
– Process connections: stainless steel, 1.4404 (316/316L); 1.4301
(304); Alloy C22, 2.4602 (UNS N06022)
Endress+Hauser 7
Page 8

Equipment architecture

2
1
6
5
7
4
3
3
Proline Promass F 300
A0027512
 1 Possibilities for integrating measuring devices into a system
1 Control system (e.g. PLC) 2 Connecting cable (0/4 to 20 mA HART etc.) 3 Fieldbus 4 Segment coupler 5 Non-hazardous area 6 Hazardous area: Zone 2; Class I, Division 2 7 Hazardous area: Zone 1; Class I, Division 1

Safety IT security

Our warranty is valid only if the device is installed and used as described in the Operating Instructions. The device is equipped with security mechanisms to protect it against any inadvertent changes to the settings.
IT security measures, which provide additional protection for the device and associated data transfer, must be implemented by the operators themselves in line with their security standards.
Device-specific IT security
The device offers a range of specific functions to support protective measures on the operator's side. These functions can be configured by the user and guarantee greater in-operation safety if used correctly. An overview of the most important functions is provided in the following section.
Function/interface Factory setting Recommendation
Write protection via hardware write protection switch →  9
Access code (also applies for Web server login or FieldCare connection) →  9
WLAN (order option in display module)
WLAN security mode Enabled (WPA2-
WLAN passphrase (password) →  9
WLAN mode Access Point On an individual basis following risk
Web server→  9 Enabled. On an individual basis following risk
CDI-RJ45 service interface →  10 On an individual basis following risk
Not enabled. On an individual basis following risk
assessment.
Not enabled (0000).
Enabled. On an individual basis following risk
PSK)
Serial number Assign a customized access code during
Assign a customized access code during commissioning.
assessment.
Do not change.
commissioning.
assessment.
assessment.
assessment.
8 Endress+Hauser
Page 9
Proline Promass F 300
Protecting access via hardware write protection
Write access to the device parameters via the local display, Web browser or operating tool (e.g. FieldCare, DeviceCare) can be disabled via a write protection switch (DIP switch on the motherboard). When hardware write protection is enabled, only read access to the parameters is possible.
Hardware write protection is disabled when the device is delivered.
Protecting access via a password
Different passwords are available to protect write access to the device parameters or access to the device via the WLAN interface.
• User-specific access code Protect write access to the device parameters via the local display, Web browser or operating tool (e.g. FieldCare, DeviceCare). Access authorization is clearly regulated through the use of a user­specific access code.
• WLAN passphrase The network key protects a connection between an operating unit (e.g. notebook or tablet) and the device via the WLAN interface which can be ordered as an option.
• Infrastructure mode When the device is operated in infrastructure mode, the WLAN passphrase corresponds to the WLAN passphrase configured on the operator side.
User-specific access code
Write access to the device parameters via the local display, Web browser or operating tool (e.g. FieldCare, DeviceCare) can be protected by the modifiable, user-specific access code.
WLAN passphrase: Operation as WLAN access point
A connection between an operating unit (e.g. notebook or tablet) and the device via the WLAN interface, which can be ordered as an optional extra, is protected by the network key. The WLAN authentication of the network key complies with the IEEE 802.11 standard.
When the device is delivered, the network key is pre-defined depending on the device. It can be changed via the WLAN settings submenu in the WLAN passphrase parameter.
Infrastructure mode
A connection between the device and WLAN access point is protected by means of an SSID and passphrase on the system side. Please contact the relevant system administrator for access.
General notes on the use of passwords
• The access code and network key supplied with the device should be changed during commissioning.
• Follow the general rules for generating a secure password when defining and managing the access code or network key.
• The user is responsible for the management and careful handling of the access code and network key.
Access via Web server
The device can be operated and configured via a Web browser with the integrated Web server. The connection is via the service interface (CDI-RJ45) or the WLAN interface. For device versions with the EtherNet/IP and PROFINET communication protocols, the connection can also be established via the terminal connection for signal transmission with EtherNet/IP or PROFINET (RJ45 connector).
The Web server is enabled when the device is delivered. The Web server can be disabled if necessary (e.g. after commissioning) via the Web server functionality parameter.
The device and status information can be hidden on the login page. This prevents unauthorized access to the information.
For detailed information on device parameters, see: The "Description of Device Parameters" document →  120
Access via OPC-UA
The "OPC UA Server" application package is available in the device version with the HART communication protocol →  117.
Endress+Hauser 9
Page 10
Proline Promass F 300
The device can communicate with OPC UA clients using the "OPC UA Server" application package.
The OPC UA server integrated in the device can be accessed via the WLAN access point using the WLAN interface - which can be ordered as an optional extra - or the service interface (CDI- RJ45) via Ethernet network. Access rights and authorization as per separate configuration.
The following Security Modes are supported as per the OPC UA Specification (IEC 62541):
• None
• Basic128Rsa15 – signed
• Basic128Rsa15 – signed and encrypted
Access via service interface (CDI-RJ45)
The device can be connected to a network via the service interface (CDI-RJ45). Device-specific functions guarantee the secure operation of the device in a network.
The use of relevant industrial standards and guidelines that have been defined by national and international safety committees, such as IEC/ISA62443 or the IEEE, is recommended. This includes organizational security measures such as the assignment of access authorization as well as technical measures such as network segmentation.
The device can be integrated in a ring topology. The device is integrated via the terminal connection for signal transmission (output 1) and the connection to the service interface (CDI­RJ45) →  105.
10 Endress+Hauser
Page 11
Proline Promass F 300

Input

Measured variable Direct measured variables

• Mass flow
• Density
• Temperature
Calculated measured variables
• Volume flow
• Corrected volume flow
• Reference density

Measuring range Measuring ranges for liquids

DN Measuring range full scale values 
min(F)
to 
[mm] [in] [kg/h] [lb/min]
8 ³⁄₈ 0 to 2 000 0 to 73.50
15 ½ 0 to 6 500 0 to 238.9
25 1 0 to 18 000 0 to 661.5
40 0 to 45 000 0 to 1 654
50 2 0 to 70 000 0 to 2 573
80 3 0 to 180 000 0 to 6 615
100 4 0 to 350 000 0 to 12 860
150 6 0 to 800 000 0 to 29 400
250 10 0 to 2 200 000 0 to 80 850
Measuring range for gases
The full scale value depends on the density and the sound velocity of the gas used and can be calculated with the formula below: 
= minimum (
max(G)
max(G)
max(F)
< 
ρ
max(G)
G
max(F)
x Constant dependent on nominal diameter
c
G
d
i
· ρG : x ; ρG · cG · π/2 · (di)2 · 3600)
max(F)
Maximum full scale value for gas [kg/h]
Maximum full scale value for liquid [kg/h]
can never be greater than 
max(G)
Gas density in [kg/m³] at operating conditions
Sound velocity (gas) [m/s]
Measuring tube internal diameter [m]
max(F)
max(F)
DN x
[mm] [in] [kg/m3]
8 ³⁄₈ 60
15 ½ 80
25 1 90
40 90
50 2 90
80 3 110
Endress+Hauser 11
Page 12
DN x
[mm] [in] [kg/m3]
100 4 130
150 6 200
250 10 200
To calculate the measuring range, use the Applicator sizing tool →  119
Calculation example for gas
• Sensor: Promass F, DN 50
• Gas: Air with a density of 60.3 kg/m³ (at 20 °C and 50 bar)
• Measuring range (liquid): 70 000 kg/h
• x = 90 kg/m³ (for Promass F, DN 50) Maximum possible full scale value:
max(G)
= 
· ρG : x = 70 000 kg/h · 60.3 kg/m³ : 90 kg/m³ = 46 900 kg/h
max(F)
Recommended measuring range
"Flow limit" section →  60
Proline Promass F 300

Operable flow range

Over 1000 : 1.
Flow rates above the preset full scale value do not override the electronics unit, with the result that the totalizer values are registered correctly.

Input signal Input and output versions

→  14
External measured values
To increase the accuracy of certain measured variables or to calculate the corrected volume flow for gases, the automation system can continuously write different measured values to the measuring device:
• Operating pressure to increase accuracy (Endress+Hauser recommends the use of a pressure measuring device for absolute pressure, e.g. Cerabar M or Cerabar S)
• Medium temperature to increase accuracy (e.g. iTEMP)
• Reference density for calculating the corrected volume flow for gases
Various pressure transmitters and temperature measuring devices can be ordered from Endress +Hauser: see "Accessories" section →  119
It is recommended to read in external measured values to calculate the corrected volume flow.
HART protocol
The measured values are written from the automation system to the measuring device via the HART protocol. The pressure transmitter must support the following protocol-specific functions:
• HART protocol
• Burst mode
Current input
The measured values are written from the automation system to the measuring device via the current input →  13.
Digital communication
The measured values can be written from the automation system to the measuring via:
• FOUNDATION Fieldbus
• PROFIBUS DP
• PROFIBUS PA
• Modbus RS485
• EtherNet/IP
• PROFINET
12 Endress+Hauser
Page 13
Proline Promass F 300
Current input 0/4 to 20 mA
Current input 0/4 to 20 mA (active/passive)
Current span • 4 to 20 mA (active)
• 0/4 to 20 mA (passive)
Resolution 1 µA
Voltage drop Typically: 0.6 to 2 V for 3.6 to 22 mA (passive)
Maximum input voltage ≤ 30 V (passive)
Open-circuit voltage ≤ 28.8 V (active)
Possible input variables • Pressure
• Temperature
• Density
Status input
Maximum input values • DC –3 to 30 V
• If status input is active (ON): Ri >3 kΩ
Response time Adjustable: 5 to 200 ms
Input signal level • Low signal: DC –3 to +5 V
• High signal: DC 12 to 30 V
Assignable functions • Off
• Reset the individual totalizers separately
• Reset all totalizers
• Flow override
Endress+Hauser 13
Page 14

Output

Proline Promass F 300

Output and input variants

Depending on the option selected for output/input 1, different options are available for the other outputs and inputs. Only one option can be selected for each output/input 1 to 3. The table must be read vertically (↓).
Example: If the option BA "4–20 mA HART" was selected for output/input 1, one of the options A, B, D, E, F, H, I or J is available for output 2 and one of the options A, B, D, E, F, H, I or J is available for output 3.
Order code for "Output; input 1" (020) → Possible options
Current output 4 to 20 mA HART BA
Current output 4 to 20 mA HART Ex i CA
FOUNDATION Fieldbus SA
FOUNDATION Fieldbus Ex i TA
PROFIBUS DP LA
PROFIBUS PA GA
PROFIBUS PA Ex i HA
Modbus RS485 MA
EtherNet/IP 2-port switch integrated NA
PROFINET 2-port switch integrated RA
Order code for "Output; input 2" (021) →
Not assigned A A A A A A A A A A
Current output 0/4 to 20 mA B B B B B B B
Current output 0/4 to 20 mA (Ex i) C C C
User configurable input/output
Pulse/frequency/switch output E E E E E E E
Double pulse output
Pulse/frequency/switch output (Ex i) G G G
Relay output H H H H H H H
Current input 0/4 to 20 mA I I I I I I I
Status input J J J J J J J
Order code for "Output; input 3" (022) →
Not assigned A A A A A A A A A A
Current output 0/4 to 20 mA B B B B B
Current output 0/4 to 20 mA (Ex i) C
User configurable input/output D D D D D
Pulse/frequency/switch output E E E E E
Double pulse output (slave) F F
Pulse/frequency/switch output (Ex i) G
Relay output H H H H H
Current input 0/4 to 20 mA I I I I I
Status input J J J J J
2)
1)
D D D D D D D
F F
1) A specific input or output can be assigned to a user configurable input/output →  18.
2) If double pulse output (F) is selected for output/input 2 (021), only the double pulse output (F) option is available for selection for output/input 3 (022).
14 Endress+Hauser
Page 15
Proline Promass F 300

Output signal HART current output

Current output 4 to 20 mA HART
Current span Can be set to: 4 to 20 mA (active/passive)
Open-circuit voltage DC 28.8 V (active)
Maximum input voltage DC 30 V (passive)
Load 250 to 700 Ω
Resolution 0.38 µA
Damping Configurable: 0.07 to 999 s
Assignable measured variables
Ex-i, passive
• Mass flow
• Volume flow
• Corrected volume flow
• Density
• Reference density
• Temperature
• Electronic temperature
• Oscillation frequency 0
• Oscillation damping 0
• Signal asymmetry
• Exciter current 0
The range of options increases if the measuring device has one or more
application packages.
PROFIBUS PA
PROFIBUS PA In accordance with EN 50170 Volume 2, IEC 61158-2 (MBP), galvanically
isolated
Data transmission 31.25 kbit/s
Current consumption 10 mA
16 mA
Permitted supply voltage 9 to 32 V
Bus connection With integrated reverse polarity protection
PROFIBUS DP
Signal encoding NRZ code
Data transfer 9.6 kBaud…12 MBaud
EtherNet/IP
Standards In accordance with IEEE 802.3
PROFINET
Standards In accordance with IEEE 802.3
FOUNDATION Fieldbus
FOUNDATION Fieldbus H1, IEC 61158-2, galvanically isolated
Data transfer 31.25 kbit/s
Endress+Hauser 15
Page 16
Current consumption 10 mA
Permitted supply voltage 9 to 32 V
Bus connection With integrated reverse polarity protection
Modbus RS485
Physical interface RS485 in accordance with EIA/TIA-485 standard
Terminating resistor Integrated, can be activated via DIP switches
Current output 0/4 to 20 mA
Current output 0/4 to 20 mA
Maximum output values 22.5 mA
Current span Can be set to:
• 4 to 20 mA (active)
• 0/4 to 20 mA (passive)
Ex-i, passive
Proline Promass F 300
Open-circuit voltage DC 28.8 V (active)
Maximum input voltage DC 30 V (passive)
Load 0 to 700 Ω
Resolution 0.38 µA
Damping Adjustable: 0.07 to 999 s
Assignable measured variables
• Mass flow
• Volume flow
• Corrected volume flow
• Density
• Reference density
• Temperature
• Electronic temperature
• Oscillation frequency 0
• Oscillation damping 0
• Signal asymmetry
• Exciter current 0
The range of options increases if the measuring device has one or more
application packages.
Pulse/frequency/switch output
Function Can be set to pulse, frequency or switch output
Version Open collector
Can be set to:
• Active
• Passive
Ex-i, passive
Maximum input values DC 30 V, 250 mA (passive)
Open-circuit voltage DC 28.8 V (active)
Voltage drop For 22.5 mA: ≤ DC 2 V
Pulse output
Maximum input values DC 30 V, 250 mA (passive)
Maximum output current 22.5 mA (active)
16 Endress+Hauser
Page 17
Proline Promass F 300
Open-circuit voltage DC 28.8 V (active)
Pulse width Adjustable: 0.05 to 2 000 ms
Maximum pulse rate 10 000 Impulse/s
Pulse value Adjustable
Assignable measured variables
Frequency output
Maximum input values DC 30 V, 250 mA (passive)
Maximum output current 22.5 mA (active)
Open-circuit voltage DC 28.8 V (active)
Output frequency Adjustable: end value frequency 2 to 10 000 Hz (f
Damping Adjustable: 0 to 999 s
Pulse/pause ratio 1:1
Assignable measured variables
Switch output
Maximum input values DC 30 V, 250 mA (passive)
Open-circuit voltage DC 28.8 V (active)
Switching behavior Binary, conductive or non-conductive
Switching delay Adjustable: 0 to 100 s
Number of switching cycles
Assignable functions • Off
• Mass flow
• Volume flow
• Corrected volume flow
• Mass flow
• Volume flow
• Corrected volume flow
• Density
• Reference density
• Temperature
• Electronic temperature
• Oscillation frequency 0
• Oscillation damping 0
• Signal asymmetry
• Exciter current 0
The range of options increases if the measuring device has one or more
application packages.
Unlimited
• On
• Diagnostic behavior
• Limit value – Mass flow – Volume flow – Corrected volume flow – Density – Reference density – Temperature – Totalizer 1-3
• Flow direction monitoring
• Status – Partially filled pipe detection – Low flow
The range of options increases if the measuring device has one or more
application packages.
= 12 500 Hz)
max
Endress+Hauser 17
Page 18
Double pulse output
Function Double pulse
Version Open collector
Can be set to:
• Active
• Passive
Maximum input values DC 30 V, 250 mA (passive)
Open-circuit voltage DC 28.8 V (active)
Voltage drop For 22.5 mA: ≤ DC 2 V
Output frequency Adjustable: 0 to 1 000 Hz
Damping Adjustable: 0 to 999 s
Pulse/pause ratio 1:1
Assignable measured variables
• Mass flow
• Volume flow
• Corrected volume flow
• Density
• Reference density
• Temperature
The range of options increases if the measuring device has one or more
application packages.
Proline Promass F 300
Relay output
Function Switch output
Version Relay output, galvanically isolated
Switching behavior Can be set to:
• NO (normally open), factory setting
• NC (normally closed)
Maximum switching capacity (passive)
Assignable functions • Off
• DC 30 V, 0.1 A
• AC 30 V, 0.5 A
• On
• Diagnostic behavior
• Limit value – Mass flow – Volume flow – Corrected volume flow – Density – Reference density – Temperature – Totalizer 1-3
• Flow direction monitoring
• Status – Partially filled pipe detection – Low flow
The range of options increases if the measuring device has one or more
application packages.
User configurable input/output
One specific input or output is assigned to a user-configurable input/output (configurable I/O)
during device commissioning.
The following inputs and outputs are available for assignment:
• Choice of current output: 4 to 20 mA (active), 0/4 to 20 mA (passive)
• Pulse/frequency/switch output
• Choice of current input: 4 to 20 mA (active), 0/4 to 20 mA (passive)
• Status input
18 Endress+Hauser
Page 19
Proline Promass F 300
The technical values correspond to those of the inputs and outputs described in this section.

Signal on alarm

Depending on the interface, failure information is displayed as follows:
HART current output
Device diagnostics Device condition can be read out via HART Command 48
PROFIBUS PA
Status and alarm messages
Error current FDE (Fault Disconnection Electronic)
Diagnostics in accordance with PROFIBUS PA Profile 3.02
0 mA
PROFIBUS DP
Status and alarm messages
Diagnostics in accordance with PROFIBUS PA Profile 3.02
EtherNet/IP
Device diagnostics Device condition can be read out in Input Assembly
PROFINET
Device diagnostics According to "Application Layer protocol for decentralized periphery", Version 2.3
FOUNDATION Fieldbus
Status and alarm messages
Error current FDE (Fault Disconnection Electronic)
Diagnostics in accordance with FF-891
0 mA
Modbus RS485
Failure mode Choose from:
• NaN value instead of current value
• Last valid value
Current output 0/4 to 20 mA
4 to 20 mA
Failure mode Choose from:
• 4 to 20 mA in accordance with NAMUR recommendation NE 43
• 4 to 20 mA in accordance with US
• Min. value: 3.59 mA
• Max. value: 22.5 mA
• Freely definable value between: 3.59 to 22.5 mA
• Actual value
• Last valid value
Endress+Hauser 19
Page 20
0 to 20 mA
Failure mode Choose from:
• Maximum alarm: 22 mA
• Freely definable value between: 0 to 20.5 mA
Pulse/frequency/switch output
Pulse output
Failure mode Choose from:
• Actual value
• No pulses
Frequency output
Failure mode Choose from:
• Actual value
• 0 Hz
• Defined value (f
Switch output
Failure mode Choose from:
• Current status
• Open
• Closed
2 to 12 500 Hz)
max
Proline Promass F 300
Relay output
Failure mode Choose from:
• Current status
• Open
• Closed
Local display
Plain text display With information on cause and remedial measures
Backlight Red backlighting indicates a device error.
Status signal as per NAMUR recommendation NE 107
Interface/protocol
• Via digital communication: – HART protocol – FOUNDATION Fieldbus – PROFIBUS PA – PROFIBUS DP – Modbus RS485 – EtherNet/IP – PROFINET
• Via service interface – CDI-RJ45 service interface – WLAN interface
Plain text display With information on cause and remedial measures
Additional information on remote operation →  101
20 Endress+Hauser
Page 21
Proline Promass F 300
Web server
Plain text display With information on cause and remedial measures
Light emitting diodes (LED)
Status information Status indicated by various light emitting diodes

Ex connection data Safety-related values

The following information is displayed depending on the device version:
• Supply voltage active
• Data transmission active
• Device alarm/error has occurred
• EtherNet/IP network available
• EtherNet/IP connection established
• PROFINET network available
• PROFINET connection established
• PROFINET blinking feature
Order code for
Output type Safety-related values
"Output; input 1"
Option BA Current output
4 to 20 mA HART
UN = 30 V UM = 250 V
Option GA PROFIBUS PA UN = 30 V
UM = 250 V
Option LA PROFIBUS DP UN = 30 V
UM = 250 V
Option MA Modbus RS485 UN = 30 V
UM = 250 V
Option SA FOUNDATION Fieldbus UN = 30 V
UM = 250 V
Option NA EtherNet/IP UN = 30 V
UM = 250 V
Option RA PROFINET UN = 30 V
UM = 250 V
Order code for "Output; input 2"; "Output; input 3"
Output type Safety-related values
Output; input 2 Output; input 3
24 (+) 25 (–) 22 (+) 23 (–)
Option B Current output
4 to 20 mA
Option D User configurable input/
output
Option E Pulse/frequency/switch
output
UN = 30 V UM = 250 V
UN = 30 V UM = 250 V
UN = 30 V UM = 250 V
Option F Double pulse output UN = 30 V
UM = 250 V
Option H Relay output UN = 30 V
IN =100 mADC/500 mA UM = 250 V
Option I Current input 4 to 20 mA UN = 30 V
UM = 250 V
Option J Status input UN = 30 V
UM = 250 V
"Output; input 1"
26 (+) 27 (–)
DC
AC
DC
AC
DC
AC
DC
AC
DC
AC
DC
AC
DC
AC
DC
AC
DC
AC
DC
AC
DC
AC
DC
AC
AC
DC
AC
DC
AC
Endress+Hauser 21
Page 22
Intrinsically safe values
Proline Promass F 300
Order code for "Output; input 1"
Option CA Current output
Option HA PROFIBUS PA Ex i Ex ia
Option TA FOUNDATION Fieldbus
1) Only available for the Zone 1; Class I, Division 1 version
2) Only available for the Zone 2; Class I, Division 2 version transmitter
Order code for "Output; input 2"; "Output; input 3"
Option C Current output
Option G Pulse/frequency/switch
Output type Intrinsically safe values
26 (+) 27 (–)
Ui = 30 V
4 to 20 mA HART Ex i
Ex i
Output type Intrinsically safe values or NIFW values
4 to 20 mA Ex i
output Ex i
li = 100 mA Pi = 1.25 W Li = 0 Ci = 0
1)
Ui = 30 V li = 570 mA Pi = 8.5 W Li = 10 µH Ci = 5 nF
1)
Ex ia
Ui = 30 V li = 570 mA Pi = 8.5 W Li = 10 µH Ci = 5 nF
Output; input 2 Output; input 3
24 (+) 25 (–) 22 (+) 23 (–)
Ui = 30 V li = 100 mA Pi = 1.25 W Li = 0 Ci = 0
Ui = 30 V li = 100 mA Pi = 1.25 W Li = 0 Ci = 0
"Output; input 1"
2)
Ex ic
Ui = 32 V li = 570 mA Pi = 8.5 W Li = 10 µH Ci = 5 nF
2)
Ex ic
Ui = 32 V li = 570 mA Pi = 8.5 W Li = 10 µH Ci = 5 nF

Low flow cut off

Galvanic isolation

The switch points for low flow cut off are user-selectable.
The outputs are galvanically isolated from one another and from earth (PE).

Protocol-specific data HART

Manufacturer ID 0x11
Device type ID 0x3B
HART protocol revision 7
Device description files (DTM, DD)
HART load Min. 250 Ω
System integration Information on system integration: Operating Instructions →  120.
Information and files under:
www.endress.com
• Measured variables via HART protocol
• Burst Mode functionality
22 Endress+Hauser
Page 23
Proline Promass F 300
PROFIBUS PA
Manufacturer ID 0x11
Ident number 0x156D
Profile version 3.02
Device description files (GSD, DTM, DD)
Supported functions • Identification & Maintenance
Configuration of the device address
Compatibility with earlier model
System integration Information regarding system integration: Operating Instructions →  120.
Information and files under:
www.endress.com
www.profibus.org
Simplest device identification on the part of the control system and nameplate
• PROFIBUS upload/download Reading and writing parameters is up to ten times faster with PROFIBUS upload/download
• Condensed status Simplest and self-explanatory diagnostic information by categorizing diagnostic messages that occur
• DIP switches on the I/O electronics module
• Local display
• Via operating tools (e.g. FieldCare)
If the device is replaced, the measuring device Promass 300 supports the compatibility of the cyclic data with previous models. It is not necessary to adjust the engineering parameters of the PROFIBUS network with the Promass 300 GSD file.
Earlier models:
• Promass 80 PROFIBUS PA – ID No.: 1528 (hex) – Extended GSD file: EH3x1528.gsd – Standard GSD file: EH3_1528.gsd
• Promass 83 PROFIBUS PA – ID No.: 152A (hex) – Extended GSD file: EH3x152A.gsd – Standard GSD file: EH3_152A.gsd
Description of the function scope of compatibility:
Operating Instructions →  120.
• Cyclic data transmission
• Block model
• Description of the modules
PROFIBUS DP
Manufacturer ID 0x11
Ident number 0x156F
Profile version 3.02
Device description files (GSD, DTM, DD)
Supported functions • Identification & Maintenance
Information and files under:
www.endress.com On the product page for the device: Documents/Software → Device drivers
www.profibus.org
Simplest device identification on the part of the control system and nameplate
• PROFIBUS upload/download Reading and writing parameters is up to ten times faster with PROFIBUS upload/download
• Condensed status Simplest and self-explanatory diagnostic information by categorizing diagnostic messages that occur
Endress+Hauser 23
Page 24
Proline Promass F 300
Configuration of the device address
System integration Information regarding system integration: Operating Instructions →  120.
• DIP switches on the I/O electronics module
• Via operating tools (e.g. FieldCare)
• Cyclic data transmission
• Block model
• Description of the modules
EtherNet/IP
Protocol • The CIP Networks Library Volume 1: Common Industrial Protocol
• The CIP Networks Library Volume 2: EtherNet/IP Adaptation of CIP
Communication type • 10Base-T
• 100Base-TX
Device profile Generic device (product type: 0x2B)
Manufacturer ID 0x11
Device type ID 0x103B
Baud rates Automatic ¹⁰⁄₁₀₀ Mbit with half-duplex and full-duplex detection
Polarity Auto-polarity for automatic correction of crossed TxD and RxD pairs
Supported CIP connections Max. 3 connections
Explicit connections Max. 6 connections
I/O connections Max. 6 connections (scanner)
Configuration options for measuring device
Configuration of the EtherNet interface
Configuration of the device address
Device Level Ring (DLR) Yes
System integration Information regarding system integration: Operating Instructions
• DIP switches on the electronics module for IP addressing
• Manufacturer-specific software (FieldCare)
• Add-on Profile Level 3 for Rockwell Automation control systems
• Web browser
• Electronic Data Sheet (EDS) integrated in the measuring device
• Speed: 10 MBit, 100 MBit, auto (factory setting)
• Duplex: half-duplex, full-duplex, auto (factory setting)
• DIP switches on the electronics module for IP addressing (last octet)
• DHCP
• Manufacturer-specific software (FieldCare)
• Add-on Profile Level 3 for Rockwell Automation control systems
• Web browser
• EtherNet/IP tools, e.g. RSLinx (Rockwell Automation)
→  120.
• Cyclic data transmission
• Block model
• Input and output groups
PROFINET
Protocol "Application layer protocol for decentral device periphery and distributed
automation", version 2.3
Communication type 100 MBit/s
Conformity class Conformance Class B
Netload Class Netload Class II
Baud rates Automatic 100 Mbit/s with full-duplex detection
Cycle times From 8 ms
Polarity Auto-polarity for automatic correction of crossed TxD and RxD pairs
24 Endress+Hauser
Page 25
Proline Promass F 300
Media Redundancy Protocol (MRP)
Device profile Application interface identifier 0xF600
Manufacturer ID 0x11
Device type ID 0x843B
Device description files (GSD, DTM, DD)
Supported connections • 1 x AR (IO Controller AR)
Configuration options for measuring device
Configuration of the device name
Supported functions • Identification & Maintenance
System integration Information regarding system integration: Operating Instructions →  120.
Yes
Generic device
Information and files under:
www.endress.com On the product page for the device: Documents/Software → Device drivers
www.profibus.org
• 1 x AR (IO-Supervisor Device AR connection allowed)
• 1 x Input CR (Communication Relation)
• 1 x Output CR (Communication Relation)
• 1 x Alarm CR (Communication Relation)
• DIP switches on the electronics module, for device name assignment (last part)
• Manufacturer-specific software (FieldCare, DeviceCare)
• Web browser
• Device master file (GSD), can be read out via the integrated Web server of the measuring device
• DIP switches on the electronics module, for device name assignment (last part)
• DCP protocol
• Process Device Manager (PDM)
• Integrated Web server
Simple device identification via: – Control system – Nameplate
• Measured value status The process variables are communicated with a measured value status
• Blinking feature via the onsite display for simple device identification and assignment
• Device operation via operating tools (e.g. FieldCare, DeviceCare, SIMATIC PDM)
• Cyclic data transmission
• Overview and description of the modules
• Status coding
• Startup configuration
• Factory setting:
FOUNDATION Fieldbus
Manufacturer ID 0x452B48 (hex)
Ident number 0x103B (hex)
Device revision 1
DD revision Information and files under:
CFF revision
Interoperability Test Kit (ITK) Version 6.2.0
ITK Test Campaign Number Information:
Link Master capability (LAS) Yes
Choice of "Link Master" and "Basic Device"
www.endress.com
www.fieldbus.org
www.endress.com
www.fieldbus.org
Yes Factory setting: Basic Device
Endress+Hauser 25
Page 26
Proline Promass F 300
Node address Factory setting: 247 (0xF7)
Supported functions The following methods are supported:
• Restart
• ENP Restart
• Diagnostic
• Set to OOS
• Set to AUTO
• Read trend data
• Read event logbook
Virtual Communication Relationships (VCRs)
Number of VCRs 44
Number of link objects in VFD 50
Permanent entries 1
Client VCRs 0
Server VCRs 10
Source VCRs 43
Sink VCRs 0
Subscriber VCRs 43
Publisher VCRs 43
Device Link Capabilities
Slot time 4
Min. delay between PDU 8
Max. response delay 16
System integration Information regarding system integration: Operating Instructions →  120.
• Cyclic data transmission
• Description of the modules
• Execution times
• Methods
Modbus RS485
Protocol Modbus Applications Protocol Specification V1.1
Response times • Direct data access: typically 25 to 50 ms
• Auto-scan buffer (data range): typically 3 to 5 ms
Device type Slave
Slave address range 1 to 247
Broadcast address range 0
Function codes • 03: Read holding register
• 04: Read input register
• 06: Write single registers
• 08: Diagnostics
• 16: Write multiple registers
• 23: Read/write multiple registers
Broadcast messages Supported by the following function codes:
• 06: Write single registers
• 16: Write multiple registers
• 23: Read/write multiple registers
26 Endress+Hauser
Page 27
Proline Promass F 300
Supported baud rate • 1 200 BAUD
• 2 400 BAUD
• 4 800 BAUD
• 9 600 BAUD
• 19 200 BAUD
• 38 400 BAUD
• 57 600 BAUD
• 115 200 BAUD
Data transfer mode • ASCII
• RTU
Data access Each device parameter can be accessed via Modbus RS485.
For Modbus register information
Compatibility with earlier model
System integration Information on system integration: Operating Instructions →  120.
If the device is replaced, the measuring device Promass 300 supports the compatibility of the Modbus registers for the process variables and the diagnostic information with the previous model Promass 83. It is not necessary to change the engineering parameters in the automation system.
Description of the function scope of compatibility:
Operating Instructions →  120.
• Modbus RS485 information
• Function codes
• Register information
• Response time
• Modbus data map

Power supply

Terminal assignment Transmitter: supply voltage, input/outputs

HART
Supply voltage Input/output 1 Input/output 2 Input/output 3
1 (+) 2 (–) 26 (+) 27 (–) 24 (+) 25 (–) 22 (+) 23 (–)
The terminal assignment depends on the specific device version ordered →  14.
FOUNDATION Fieldbus
Supply voltage Input/output 1 Input/output 2 Input/output 3
1 (+) 2 (–) 26 (A) 27 (B) 24 (+) 25 (–) 22 (+) 23 (–)
The terminal assignment depends on the specific device version ordered →  14.
PROFIBUS PA
Supply voltage Input/output 1 Input/output 2 Input/output 3
1 (+) 2 (–) 26 (B) 27 (A) 24 (+) 25 (–) 22 (+) 23 (–)
The terminal assignment depends on the specific device version ordered →  14.
PROFIBUS DP
Supply voltage Input/output 1 Input/output 2 Input/output 3
1 (+) 2 (–) 26 (B) 27 (A) 24 (+) 25 (–) 22 (+) 23 (–)
The terminal assignment depends on the specific device version ordered →  14.
Endress+Hauser 27
Page 28
Proline Promass F 300
Modbus RS485
Supply voltage Input/output 1 Input/output 2 Input/output 3
1 (+) 2 (–) 26 (B) 27 (A) 24 (+) 25 (–) 22 (+) 23 (–)
The terminal assignment depends on the specific device version ordered →  14.
PROFINET
Supply voltage Input/output 1 Input/output 2 Input/output 3
1 (+) 2 (–) PROFINET
(RJ45 connector)
EtherNet/IP
Supply voltage Input/output 1 Input/output 2 Input/output 3
1 (+) 2 (–) EtherNet/IP
(RJ45 connector)
24 (+) 25 (–) 22 (+) 23 (–)
The terminal assignment depends on the specific
device version ordered →  14.
24 (+) 25 (–) 22 (+) 23 (–)
The terminal assignment depends on the specific
device version ordered →  14.

Device plugs available

Terminal assignment of the remote display and operating module →  33.
Device plugs may not be used in hazardous areas!
Device plugs for fieldbus systems:
Order code for "Input; output 1"
• Option SA "FOUNDATION Fieldbus" →  28
• Option GA "PROFIBUS PA" →  28
• Option RA "PROFINET" →  29
• Option NA "EtherNet/IP" →  29
Device plug for connecting to the service interface:
Order code for "Accessory mounted" option NB, adapter RJ45 M12 (service interface) →  30
Order code for "Input; output 1", option SA "FOUNDATION Fieldbus"
Order code for Cable entry/connection →  32
"Electrical connection" 2 3
M, 3, 4, 5 7/8" connector
Order code for "Input; output 1", option GA "PROFIBUS PA"
Order code for Cable entry/connection →  32
"Electrical connection" 2 3
L, N, P, U Connector M12 × 1
28 Endress+Hauser
Page 29
Proline Promass F 300
1
2
4
3
1
2
4
3
Order code for "Input; output 1", option RA "PROFINET"
Order code for Cable entry/connection →  32
"Electrical connection" 2 3
L, N, P, U Connector M12 × 1
1) 2)
1) 2)
1) 2)
R
, S
, T
, V
1) 2)
Connector M12 × 1 Connector M12 × 1
1) Cannot be combined with an external WLAN antenna (order code for "Enclosed accessories", option P8) of an RJ45 M12 adapter for the service interface (order code for "Accessories mounted", option NB) or of the remote display and operating module DKX001.
2) Suitable for integrating the device in a ring topology.
Order code for "Input; output 1", option NA "EtherNet/IP"
Order code for Cable entry/connection →  32
"Electrical connection" 2 3
L, N, P, U Connector M12 × 1
1) 2)
1) 2)
1) 2)
R
, S
, T
, V
1) 2)
Connector M12 × 1 Connector M12 × 1
1) Cannot be combined with an external WLAN antenna (order code for "Enclosed accessories", option P8) of an RJ45 M12 adapter for the service interface (order code for "Accessories mounted", option NB) or of the remote display and operating module DKX001
2) Suitable for integrating the device in a ring topology.
Order code for "Accessory mounted", option NB "Adapter RJ45 M12 (service interface)"
Order code Cable entry/coupling →  32
"Accessory mounted" Cable entry
NB Plug M12 × 1

Pin assignment, device plug FOUNDATION Fieldbus

PROFIBUS PA
Cable entry
2
3
Pin Assignment Coding Plug/socket
1 + Signal + A Plug
2 - Signal –
3 Grounding
4 Not assigned
Pin Assignment Coding Plug/socket
1 + PROFIBUS PA + A Plug
2 Grounding
3 - PROFIBUS PA –
4 Not assigned
Endress+Hauser 29
Page 30
PROFINET
3
2
4
3
2
4
3
2
4
Pin Assignment
1 + TD +
2 + RD +
3 - TD –
4 - RD –
A0032047
Recommended plug:
• Binder, series 763, part no. 99 3729 810 04
• Phoenix, part no. 1543223 SACC-M12MSD-4Q
• When using the device in a hazardous location, use a suitably certified plug.
Coding Plug/socket
D Socket
EtherNet/IP
Pin Assignment
1 + Tx
2 + Rx
3 - Tx
4 - Rx
A0032047
Coding Plug/socket
D Socket
Proline Promass F 300

Supply voltage

Recommended plug:
• Binder, series 763, part no. 99 3729 810 04
• Phoenix, part no. 1543223 SACC-M12MSD-4Q
• When using the device in a hazardous location, use a suitably certified plug.
Service interface
Order code for "Accessories mounted", option NB: Adapter RJ45 M12 (service interface)
Pin Assignment
1 + Tx
2 + Rx
3 - Tx
4 - Rx
A0032047
Recommended plug:
• Binder, series 763, part no. 99 3729 810 04
• Phoenix, part no. 1543223 SACC-M12MSD-4Q
• When using the device in a hazardous location, use a suitably certified plug.
Order code for "Power supply"
Option D DC24 V ±20%
Option E AC100 to 240 V –15…+10% 50/60 Hz
Option I
Coding Plug/socket
D Socket
terminal voltage Frequency range
DC24 V ±20%
AC100 to 240 V –15…+10% 50/60 Hz
30 Endress+Hauser
Page 31
Proline Promass F 300

Power consumption Transmitter

Max. 10 W (active power)

Current consumption Transmitter

• Max. 400 mA (24 V)
• Max. 200 mA (110 V, 50/60 Hz; 230 V, 50/60 Hz)

Power supply failure

Depending on the device version, the configuration is retained in the device memoryor in the pluggable data memory (HistoROM DAT).
Endress+Hauser 31
Page 32

Electrical connection Connecting the transmitter

1
2
3 4
1
2
3 4
• Terminal assignment→  27
• Device plugs available→  28
1 Terminal connection for supply voltage 2 Terminal connection for signal transmission, input/output 3 Terminal connection for signal transmission, input/output or terminal for network connection via service
interface (CDI-RJ45); Optional: terminal connection for external WLAN antenna or connection for remote display and operating module DKX001
4 Protective ground (PE)
An adapter for RJ45 and the M12 connector is optionally available: Order code for "Accessories", option NB: "Adapter RJ45 M12 (service interface)"
The adapter connects the service interface (CDI-RJ45) to an M12 connector mounted in the cable entry. Therefore the connection to the service interface can be established via an M12 connector without opening the device.
Network connection via service interface (CDI-RJ45) →  107
Proline Promass F 300
A0026781
Connecting in a ring topology
Device versions with EtherNet/IP and PROFINET communication protocols can be integrated into a ring topology. The device is integrated via the terminal connection for signal transmission (output 1) and the connection to the service interface (CDI-RJ45).
Integrate the transmitter into a ring topology:
• EtherNet/IP →  105
• PROFINET →  106
A0026781
1 Terminal connection for supply voltage 2 Terminal connection for signal transmission: PROFINET or EtherNet/IP (RJ45 connector) 3 Connection to service interface (CDI-RJ45) 4 Protective earth (PE)
If the device has additional input/outputs, these are routed via the cable entry for the connection to the service interface (CDI-RJ45).
32 Endress+Hauser
Page 33
Proline Promass F 300
1
3
81
Vcc
82
Gnd
83A84
B
81
Vcc
82
Gnd
83A84
B
2
4 5
4
4...20 mA
5
2
1
3
6
Connecting the remote display and operating module DKX001
The remote display and operating module DKX001 is available as an optional extra →  117.
• The remote display and operating module DKX001 is only available for the following housing versions, order code for "Housing": – Option A "Aluminum, coated" – Option L "Cast, stainless"
• The measuring device is always supplied with a dummy cover when the remote display and operating module DKX001 is ordered directly with the measuring device. Display or operation at the transmitter is not possible in this case.
• If ordered subsequently, the remote display and operating module DKX001 may not be connected at the same time as the existing measuring device display module. Only one display or operation unit may be connected to the transmitter at any one time.
A0027518
1 Remote display and operating module DKX001 2 Protective earth (PE) 3 Connecting cable 4 Measuring device 5 Protective earth (PE)
Connection examples
Current output 4 to 20 mA HART
A0029055
 2 Connection example for 4 to 20 mA HART current output (active)
1 Automation system with current input (e.g. PLC) 2 Cable shield: the cable shield must be grounded at both ends to comply with EMC requirements; observe cable
specifications →  41 3 Connection for HART operating devices →  101 4 Resistor for HART communication (≥ 250 Ω): observe maximum load →  15 5 Analog display unit: observe maximum load →  15 6 Transmitter
Endress+Hauser 33
Page 34
Proline Promass F 300
2
3
4...20 mA
41
5
2
4...20 mA
4
1
2
3
3
6
5
A0028762
 3 Connection example for 4 to 20 mA HART current output (passive)
1 Automation system with current input (e.g. PLC) 2 Power supply 3 Cable shield: the cable shield must be grounded at both ends to comply with EMC requirements; observe cable
specifications →  41 4 Analog display unit: observe maximum load →  15 5 Transmitter
HART input
A0028763
 4 Connection example for HART input with a common negative (passive)
1 Automation system with HART output (e.g. PLC) 2 Active barrier for power supply (e.g. RN221N) 3 Cable shield: the cable shield must be grounded at both ends to comply with EMC requirements; observe cable
specifications 4 Analog display unit: observe maximum load 5 Pressure transmitter (e.g. Cerabar M, Cerabar S): see requirements 6 Transmitter
34 Endress+Hauser
Page 35
Proline Promass F 300
21 3 4
78
6 6
6
6
5
6
6
5
2
1
A
B
3
4
4
A
B
A
B
PROFIBUS PA
A0028768
 5 Connection example for PROFIBUS PA
1 Control system (e.g. PLC) 2 PROFIBUS PA segment coupler 3 Cable shield: the cable shield must be grounded at both ends to comply with EMC requirements; observe cable
specifications 4 T-box 5 Measuring device 6 Local grounding 7 Bus terminator 8 Potential matching line
PROFIBUS DP
A0028765
 6 Connection example for PROFIBUS DP, non-hazardous area and Zone 2/Div. 2
1 Control system (e.g. PLC) 2 Cable shield: the cable shield must be grounded at both ends to comply with EMC requirements; observe cable
specifications 3 Distribution box 4 Transmitter
If baud rates > 1.5 MBaud an EMC cable entry must be used and the cable shield must continue
Endress+Hauser 35
as far as the terminal wherever possible.
Page 36
EtherNet/IP
1
2
4
3
5
5
1 2 4
5
5
3
1 2 4
5
5
3
 7 Connection example for EtherNet/IP
1 Control system (e.g. PLC) 2 Ethernet switch 3 Observe cable specifications 4 Device plug 5 Transmitter
EtherNet/IP: DLR (Device Level Ring)
Proline Promass F 300
A0028767
1 Control system (e.g. PLC) 2 Ethernet switch 3 Observe cable specifications →  41 4 Connecting cable between the two transmitters 5 Transmitter
PROFINET
 8 Connection example for PROFINET
1 Control system (e.g. PLC) 2 Ethernet switch 3 Observe cable specifications 4 Device plug 5 Transmitter
A0027544
A0016805
36 Endress+Hauser
Page 37
Proline Promass F 300
1 2 4
5
5
3
21 3 4
78
6 6
6
6
5
6
6
5
PROFINET: MRP (Media Redundancy Protocol)
A0027544
1 Control system (e.g. PLC) 2 Ethernet switch 3 Observe cable specifications →  41 4 Connecting cable between the two transmitters 5 Transmitter
FOUNDATION Fieldbus
 9 Connection example for FOUNDATION Fieldbus
1 Control system (e.g. PLC) 2 Power Conditioner (FOUNDATION Fieldbus) 3 Cable shield: the cable shield must be grounded at both ends to comply with EMC requirements; observe cable
specifications 4 T-box 5 Measuring device 6 Local grounding 7 Bus terminator 8 Potential matching line
Endress+Hauser 37
A0028768
Page 38
Proline Promass F 300
2
1
A
B
3
4
4
A
B
A
B
4...20 mA
2
1
3
2
4...20 mA
3
1
4
Modbus RS485
A0028765
 10 Connection example for Modbus RS485, non-hazardous area and Zone 2; Class I, Division 2
1 Control system (e.g. PLC) 2 Cable shield: the cable shield must be grounded at both ends to comply with EMC requirements; observe cable
specifications 3 Distribution box 4 Transmitter
Current output 4-20 mA
A0028758
 11 Connection example for 4-20 mA current output (active)
1 Automation system with current input (e.g. PLC) 2 Analog display unit: observe maximum load 3 Transmitter
A0028759
 12 Connection example for 4-20 mA current output (passive)
1 Automation system with current input (e.g. PLC) 2 Active barrier for power supply (e.g. RN221N) 3 Analog display unit: observe maximum load 4 Transmitter
38 Endress+Hauser
Page 39
Proline Promass F 300
1
2
3
12345
1
2
3
1
2
3
4
Pulse/frequency output
A0028761
 13 Connection example for pulse/frequency output (passive)
1 Automation system with pulse/frequency input (e.g. PLC) 2 Power supply 3 Transmitter: Observe input values →  16
Switch output
 14 Connection example for switch output (passive)
1 Automation system with switch input (e.g. PLC) 2 Power supply 3 Transmitter: Observe input values →  16
Double pulse output
 15 Connection example for double pulse output (active)
1 Automation system with double pulse input (e.g. PLC) 2 Transmitter: Observe input values →  18 3 Double pulse output 4 Double pulse output (slave), phase-shifted
A0028760
A0029280
Endress+Hauser 39
Page 40
1
3
2
4
5
 16 Connection example for double pulse output (passive)
1
2
3
31
4
2
1 Automation system with double pulse input (e.g. PLC) 2 Power supply 3 Transmitter: Observe input values →  18 4 Double pulse output 5 Double pulse output (slave), phase-shifted
Proline Promass F 300
A0029279
Relay output
 17 Connection example for relay output (passive)
1 Automation system with relay input (e.g. PLC) 2 Power supply 3 Transmitter: Observe input values →  18
Current input
A0028760
A0028915
 18 Connection example for 4 to 20 mA current input
1 Power supply 2 Terminal box 3 External measuring device (for reading in pressure or temperature, for instance) 4 Transmitter
40 Endress+Hauser
Page 41
Proline Promass F 300
1
2
3
Status input
 19 Connection example for status input
1 Automation system with status output (e.g. PLC) 2 Power supply 3 Transmitter

Potential equalization Requirements

No special measures for potential equalization are required.
Please consider the following to ensure correct measurement:
• Same electrical potential for the fluid and sensor
• Company-internal grounding concepts
A0028764

terminals

Spring-loaded terminals: Suitable for strands and strands with ferrules. Conductor cross-section 0.2 to 2.5 mm2 (24 to 12 AWG).

Cable entries

• Cable gland: M20 × 1.5 with cable ⌀ 6 to 12 mm (0.24 to 0.47 in)
• Thread for cable entry: – NPT ½" – G ½" – M20
• Device plug for digital communication: M12 Only available for certain device versions →  28.

Cable specification Permitted temperature range

• The installation guidelines that apply in the country of installation must be observed.
• The cables must be suitable for the minimum and maximum temperatures to be expected.
Power supply cable
Standard installation cable is sufficient.
Signal cable
Current output 4 to 20 mA HART
A shielded cable is recommended. Observe grounding concept of the plant.
PROFIBUS PA
Twisted, shielded two-wire cable. Cable type A is recommended .
For further information on planning and installing PROFIBUS networks see:
• Operating Instructions "PROFIBUS DP/PA: Guidelines for planning and commissioning" (BA00034S)
• PNO Directive 2.092 "PROFIBUS PA User and Installation Guideline"
• IEC 61158-2 (MBP)
PROFIBUS DP
The IEC 61158 standard specifies two types of cable (A and B) for the bus line which can be used for every transmission rate. Cable type A is recommended.
Endress+Hauser 41
Page 42
Proline Promass F 300
Cable type A
Characteristic impedance 135 to 165 Ω at a measuring frequency of 3 to 20 MHz
Cable capacitance < 30 pF/m
Wire cross-section
Cable type Twisted pairs
Loop resistance ≤110 Ω/km
Signal damping Max. 9 dB over the entire length of the cable cross-section
Shield Copper braided shielding or braided shielding with foil shield. When grounding
For further information on planning and installing PROFIBUS networks see:
• Operating Instructions "PROFIBUS DP/PA: Guidelines for planning and commissioning" (BA00034S)
• PNO Directive 2.092 "PROFIBUS PA User and Installation Guideline"
• IEC 61158-2 (MBP)
EtherNet/IP
The standard ANSI/TIA/EIA-568-B.2 Annex specifies CAT 5 as the minimum category for a cable used for EtherNet/IP. CAT 5e and CAT 6 are recommended.
For more information on planning and installing EtherNet/IP networks, please refer to the "Media Planning and Installation Manual. EtherNet/IP" of ODVA Organization
> 0.34 mm2 (22 AWG)
the cable shield, observe the grounding concept of the plant.
PROFINET
Standard IEC 61156-6 specifies CAT 5 as the minimum category for a cable used for PROFINET. CAT 5e and CAT 6 are recommended.
For more information on planning and installing PROFINET networks, see: "PROFINET Cabling and Interconnection Technology", Guideline for PROFINET
FOUNDATION Fieldbus
Twisted, shielded two-wire cable.
For further information on planning and installing FOUNDATION Fieldbus networks see:
• Operating Instructions for "FOUNDATION Fieldbus Overview" (BA00013S)
• FOUNDATION Fieldbus Guideline
• IEC 61158-2 (MBP)
Modbus RS485
The EIA/TIA-485 standard specifies two types of cable (A and B) for the bus line which can be used for every transmission rate. Cable type A is recommended.
Cable type A
Characteristic impedance 135 to 165 Ω at a measuring frequency of 3 to 20 MHz
Cable capacitance < 30 pF/m
Wire cross-section
Cable type Twisted pairs
Loop resistance ≤110 Ω/km
Signal damping Max. 9 dB over the entire length of the cable cross-section
Shield Copper braided shielding or braided shielding with foil shield. When grounding
> 0.34 mm2 (22 AWG)
the cable shield, observe the grounding concept of the plant.
Current output 0/4 to 20 mA
Standard installation cable is sufficient.
42 Endress+Hauser
Page 43
Proline Promass F 300
Pulse/frequency/switch output
Standard installation cable is sufficient.
Double pulse output
Standard installation cable is sufficient.
Relay output
Standard installation cable is sufficient.
Current input 0/4 to 20 mA
Standard installation cable is sufficient.
Status input
Standard installation cable is sufficient.
Connecting cable for transmitter - remote display and operating module DKX001
Standard cable
A standard cable can be used as the connecting cable.
Standard cable 4 cores (2 pairs); pair-stranded with common shield
Shielding Tin-plated copper-braid, optical cover ≥ 85 %
Capacitance: core/shield Maximum 1 000 nF for Zone 1; Class I, Division 1
L/R Maximum 24 µH/Ω for Zone 1; Class I, Division 1
Cable length Maximum 300 m (1 000 ft), see the following table
Cable length for use in:
Cross-section
0.34 mm2 (22 AWG) 80 m (270 ft)
0.50 mm2 (20 AWG) 120 m (400 ft)
0.75 mm2 (18 AWG) 180 m (600 ft)
1.00 mm2 (17 AWG) 240 m (800 ft)
1.50 mm2 (15 AWG) 300 m (1 000 ft)
Non-hazardous area
Hazardous area: Zone 2; Class I, Division 2
Hazardous area: Zone 1; Class I, Division 1
Optionally available connecting cable
Standard cable
Flame resistance According to DIN EN 60332-1-2
Oil-resistance According to DIN EN 60811-2-1
Shielding Tin-plated copper-braid, optical cover ≥ 85 %
Capacitance: core/shield ≤200 pF/m
L/R ≤24 µH/Ω
Available cable length 10 m (35 ft)
Operating temperature When mounted in a fixed position: –50 to +105 °C (–58 to +221 °F); when cable
2 × 2 × 0.34 mm2 (22 AWG) PVC cable stranded)
can move freely: –25 to +105 °C (–13 to +221 °F)
1)
with common shield (2 pairs, pair-
1) UV radiation can impair the cable outer sheath. Protect the cable from direct sunshine where possible.
Endress+Hauser 43
Page 44

Performance characteristics

Proline Promass F 300

Reference operating conditions

Maximum measured error

• Error limits based on ISO 11631
• Water with +15 to +45 °C (+59 to +113 °F) at2 to 6 bar (29 to 87 psi)
• Specifications as per calibration protocol
• Accuracy based on accredited calibration rigs that are traced to ISO 17025. To obtain measured errors, use the Applicator sizing tool →  119
o.r. = of reading; 1 g/cm³ = 1 kg/l; T = medium temperature
Base accuracy
Design fundamentals →  48
Mass flow and volume flow (liquids)
±0.05 % o.r. (PremiumCal; order code for "Calibration flow", option D, for mass flow) ±0.10 % o.r.
Mass flow (cryogenic liquids)
Order code for "Measuring tube material", option LA
±0.35 % o.r.
Mass flow (gases)
±0.35 % o.r.
Density (liquids)
Under
reference operating conditions
[g/cm³] [g/cm³] [g/cm³]
±0.0005 ±0.01 ±0.001
1) Valid over the entire temperature and density range
2) Valid range for special density calibration: 0 to 2 g/cm³, +5 to +80 °C (+41 to +176 °F)
3) Order code for "Application package", option EE "Special density" only in combination with the order code
for "Measuring tube mat., wetted surface", option BB, BF, HA, SA
Standard density
calibration
1)
Wide-range
Density specification
2) 3)
Density (cryogenic liquids)
Order code for "Measuring tube material", option LA
±0.05 g/cm³
Temperature
±0.5 °C ± 0.005 · T °C (±0.9 °F ± 0.003 · (T – 32) °F)
Zero point stability
DN Zero point stability
[mm] [in] [kg/h] [lb/min]
8 ³⁄₈ 0.030 0.001
15 ½ 0.200 0.007
25 1 0.540 0.019
40 2.25 0.083
50 2 3.50 0.129
44 Endress+Hauser
Page 45
Proline Promass F 300
DN Zero point stability
[mm] [in] [kg/h] [lb/min]
80 3 9.0 0.330
100 4 14.0 0.514
150 6 32.0 1.17
250 10 88.0 3.23
High-temperature version: order code for "Measuring tube material", option TT, TU
DN Zero point stability
[mm] [in] [kg/h] [lb/min]
25 1 1.80 0.0661
50 2 7.00 0.2572
80 3 18.0 0.6610
For devices with low-temperature version, order code for "Measuring tube mat., wetted surface", option LA, please note the following:
NOTICE
Zero point confirmation and zero point adjustment are difficult to carry out in the field due to the vaporization of the cryogenic liquid.
As a general rule, the factory-set zero point should not be changed. Please ensure that the
medium is in the liquid phase if a zero point adjustment is to be carried out.
Flow values
Flow values as turndown parameter depending on nominal diameter.
SI units
DN 1:1 1:10 1:20 1:50 1:100 1:500
[mm]
8 2 000 200 100 40 20 4
15 6 500 650 325 130 65 13
25 18 000 1 800 900 360 180 36
40 45 000 4 500 2 250 900 450 90
50 70 000 7 000 3 500 1 400 700 140
80 180 000 18 000 9 000 3 600 1 800 360
100 350 000 35 000 17 500 7 000 3 500 700
150 800 000 80 000 40 000 16 000 8 000 1 600
250 2 200 000 220 000 110 000 44 000 22 000 4 400
[kg/h] [kg/h] [kg/h] [kg/h] [kg/h] [kg/h]
US units
DN 1:1 1:10 1:20 1:50 1:100 1:500
[inch] [lb/min] [lb/min] [lb/min] [lb/min] [lb/min] [lb/min]
³⁄₈ 73.50 7.350 3.675 1.470 0.735 0.147
½ 238.9 23.89 11.95 4.778 2.389 0.478
1 661.5 66.15 33.08 13.23 6.615 1.323
1 654 165.4 82.70 33.08 16.54 3.308
Endress+Hauser 45
Page 46
Proline Promass F 300
DN 1:1 1:10 1:20 1:50 1:100 1:500
[inch] [lb/min] [lb/min] [lb/min] [lb/min] [lb/min] [lb/min]
2 2 573 257.3 128.7 51.46 25.73 5.146
3 6 615 661.5 330.8 132.3 66.15 13.23
4 12 860 1 286 643.0 257.2 128.6 25.72
6 29 400 2 940 1 470 588 294 58.80
10 80 850 8 085 4 043 1 617 808.5 161.7
Accuracy of outputs
The outputs have the following base accuracy specifications.
Current output
Accuracy ±5 µA
Pulse/frequency output
o.r. = of reading

Repeatability

Accuracy Max. ±50 ppm o.r. (over the entire ambient temperature range)
o.r. = of reading; 1 g/cm3 = 1 kg/l; T = medium temperature
Base repeatability
Design fundamentals →  48
Mass flow and volume flow (liquids)
±0.025 % o.r. (PremiumCal, for mass flow) ±0.05 % o.r.
Mass flow (cryogenic liquids)
Order code for "Measuring tube material", option LA ±0.175 % % o.r.
Mass flow (gases)
±0.25 % o.r.
Density (liquids)
±0.00025 g/cm
3
Density (cryogenic liquids)
Order code for "Measuring tube material", option LA ±0.025 g/cm
3
Temperature
±0.25 °C ± 0.0025 · T °C (±0.45 °F ± 0.0015 · (T–32) °F)

Response time

Influence of ambient
The response time depends on the configuration (damping).
Current output
temperature
Temperature coefficient Max. 1 μA/°C
46 Endress+Hauser
Page 47
Proline Promass F 300
3
[°C]
0- 08 80 160 240
[° ]F
- 05 0
100
200
300
0
2
4
6
8
10
2
1
12
14
16
18
320 400 480 560
640
Pulse/frequency output
Temperature coefficient No additional effect. Included in accuracy.

Influence of medium temperature

Mass flow and volume flow
o.f.s. = of full scale value
When there is a difference between the temperature for zero point adjustment and the process temperature, the additional measured error of the sensor is typically ±0.0002 % o.f.s./°C (±0.0001 % o. f.s./°F).
The effect is reduced if zero point adjustment is performed at process temperature.
Density
When there is a difference between the density calibration temperature and the process temperature, the typical measured error of the sensor is ±0.00005 g/cm3 /°C (±0.000025 g/cm3 /°F). Field density calibration is possible.
Wide-range density specification (special density calibration)
If the process temperature is outside the valid range (→  44) the measured error is ±0.00005 g/cm3 /°C (±0.000025 g/cm3 /°F)

Influence of medium pressure

Endress+Hauser 47
A0027453
1 Field density calibration, for example at +20 °C (+68 °F) 2 Special density calibration
Temperature
±0.005 · T °C (± 0.005 · (T – 32) °F)
The table below shows the effect on accuracy of mass flow due to a difference between calibration pressure and process pressure.
o.r. = of reading
It is possible to compensate for the effect by:
• Reading in the current pressure measured value via the current input.
• Specifying a fixed value for the pressure in the device parameters. Operating Instructions→  120.
DN [% o.r./bar] [% o.r./psi]
[mm] [in]
8 3/8 no influence
15 ½ no influence
Page 48
Proline Promass F 300
ZeroPoint BaseAccu
100
³
± BaseAccu
ZeroPoint BaseAccu
100
<
ZeroPoint
MeasValue
100
±
½ ZeroPoint
BaseRepeat
100
³
½ ZeroPoint
BaseRepeat
100
<
ZeroPoint
MeasValue
100
½ ±
E [%]
0 10 20 30 40 50 60
0
0.5
1.0
1.5
2.0
2.5
70 80 90 100
Q [%]
DN [% o.r./bar] [% o.r./psi]
[mm] [in]
25 1 no influence
40 –0.003 –0.0002
50 2 –0.008 –0.0006
80 3 –0.009 –0.0006
100 4 –0.007 –0.0005
150 6 –0.009 –0.0006
250 10 –0.009 –0.0006

Design fundamentals

o.r. = of reading, o.f.s. = of full scale value
BaseAccu = base accuracy in % o.r., BaseRepeat = base repeatability in % o.r.
MeasValue = measured value; ZeroPoint = zero point stability
Calculation of the maximum measured error as a function of the flow rate
Flow rate Maximum measured error in % o.r.
A0021332
A0021333
Calculation of the maximum repeatability as a function of the flow rate
Flow rate Maximum repeatability in % o.r.
A0021335
A0021336
A0021339
A0021334
A0021340
A0021337
Example for maximum measured error
E Maximum measured error in % o.r. (example with PremiumCal) Q Flow rate in % of maximum full scale value

Installation

No special measures such as supports etc. are necessary. External forces are absorbed by the construction of the device.
48 Endress+Hauser
A0028808
Page 49
Proline Promass F 300
1
2
3
4
5

Mounting location

A0028772
To prevent measuring errors arising from accumulation of gas bubbles in the measuring tube, avoid the following mounting locations in the pipe:
• Highest point of a pipeline.
• Directly upstream of a free pipe outlet in a down pipe.
Installation in down pipes
However, the following installation suggestion allows for installation in an open vertical pipeline. Pipe restrictions or the use of an orifice with a smaller cross-section than the nominal diameter prevent the sensor running empty while measurement is in progress.
A0028773
 20 Installation in a down pipe (e.g. for batching applications)
1 Supply tank 2 Sensor 3 Orifice plate, pipe restriction 4 Valve 5 Batching tank
DN Ø orifice plate, pipe restriction
[mm] [in] [mm] [in]
8 ³⁄₈ 6 0.24
15 ½ 10 0.40
25 1 14 0.55
40 22 0.87
50 2 28 1.10
80 3 50 1.97
100 4 65 2.60
Endress+Hauser 49
Page 50
Proline Promass F 300
1 2
DN Ø orifice plate, pipe restriction
[mm] [in] [mm] [in]
150 6 90 3.54
250 10 150 5.91

Orientation

The direction of the arrow on the sensor nameplate helps you to install the sensor according to the flow direction (direction of medium flow through the piping).
Orientation Recommendation
A Vertical orientation
A0015591
B Horizontal orientation, transmitter at
top
→  21,  50
A0015589
C Horizontal orientation, transmitter at
bottom
→  21,  50
A0015590
D Horizontal orientation, transmitter at
side
A0015592
1) Applications with low process temperatures may decrease the ambient temperature. To maintain the
minimum ambient temperature for the transmitter, this orientation is recommended.
2) Applications with high process temperatures may increase the ambient temperature. To maintain the
maximum ambient temperature for the transmitter, this orientation is recommended.
1)
Exceptions:
2)
Exceptions:

Inlet and outlet runs

Special mounting instructions

If a sensor is installed horizontally with a curved measuring tube, match the position of the sensor to the fluid properties.
A0028774
 21 Orientation of sensor with curved measuring tube
1 Avoid this orientation for fluids with entrained solids: Risk of solids accumulating. 2 Avoid this orientation for outgassing fluids: Risk of gas accumulating.
No special precautions need to be taken for fittings which create turbulence, such as valves, elbows or T-pieces, as long as no cavitation occurs →  60.
Rupture disk
Information that is relevant to the process: →  60.
50 Endress+Hauser
Page 51
Proline Promass F 300
RUPTURE DISK
i
2
3
1
DN 8 ( ")...150 (6")!⁄# DN 250 (10")
2
3
WARNING
L
Danger from medium escaping!
Medium escaping under pressure can cause injury or material damage.
Take precautions to prevent danger to persons and damage if the rupture disk is actuated.
Observe information on the rupture disk sticker.
Make sure that the function and operation of the rupture disk is not impeded through the
installation of the device. Do not use a heating jacket.
Do not remove or damage the rupture disk.
The position of the rupture disk is indicated on a sticker beside it.
The transportation guard must be removed.
The existing connecting nozzles are not intended for the purpose of rinsing or pressure monitoring, but instead serve as the mounting location for the rupture disk.
In the event of a failure of the rupture disk, a discharge device can be screwed onto the internal thread of the rupture disk in order to drain off any escaping medium.
1 Rupture disk label 2 Rupture disk with 1/2" NPT internal thread with 1" width across flat 3 Transport protection
For information on the dimensions: see the "Mechanical construction" section (accessories)
Zero point adjustment
All measuring devices are calibrated in accordance with state-of-the-art technology. Calibration takes place under reference conditions→  44. Therefore, a zero point adjustment in the field is generally not required.
Experience shows that zero point adjustment is advisable only in special cases:
• To achieve maximum measuring accuracy even with low flow rates.
• Under extreme process or operating conditions (e.g. very high process temperatures or very high­viscosity fluids).
A0028903
Endress+Hauser 51
Page 52

Environment

Proline Promass F 300

Ambient temperature range

Storage temperature

Climate class

Degree of protection

Measuring device • –40 to +60 °C (–40 to +140 °F)
• Order code for "Test, certificate", option JP: –50 to +60 °C (–58 to +140 °F)
Readability of the local display
Dependency of ambient temperature on medium temperature→  53
If operating outdoors:
Avoid direct sunlight, particularly in warm climatic regions.
You can order a weather protection cover from Endress+Hauser. →  117.
–50 to +80 °C (–58 to +176 °F)
DIN EN 60068-2-38 (test Z/AD)
Measuring device
• As standard: IP66/67, type 4X enclosure
• When housing is open: IP20, type 1 enclosure
• Display module: IP20, type 1 enclosure
• With the order code for "Sensor options", option CM: IP69 can also be ordered
External WLAN antenna
IP67
–20 to +60 °C (–4 to +140 °F) The readability of the display may be impaired at temperatures outside the temperature range.

Vibration resistance

Shock resistance

Shock resistance

Interior cleaning

Electromagnetic compatibility (EMC)

• Oscillation, sinusoidal, following IEC 60068-2-6 – 2 to 8.4 Hz, 3.5 mm peak – 8.4 to 2 000 Hz, 1 g peak
• Oscillation, broadband noise following IEC 60068-2-64 – 10 to 200 Hz, 0.003 g2/Hz – 200 to 2 000 Hz, 0.001 g2/Hz – Total: 1.54 g rms
Shock, half-sine according to IEC 60068-2-27 6 ms 50 g
Shock due to rough handling following IEC 60068-2-31
• Cleaning in place (CIP)
• Sterilization in place (SIP)
Options
• Oil- and grease-free version for wetted parts, without declaration Order code for "Service", option HA
• Oil- and grease-free version for wetted parts as per IEC/TR 60877-2.0 and BOC 50000810-4, with declaration Order code for "Service", option HB
• As per IEC/EN 61326 and NAMUR Recommendation 21 (NE 21)
• Device version with PROFIBUS DP: Complies with emission limits for industry as per EN 50170 Volume 2, IEC 61784
The following applies for PROFIBUS DP: If baud rates > 1.5 MBaud, an EMC cable entry must be used and the cable shield must continue as far as the terminal wherever possible.
Details are provided in the Declaration of Conformity.
52 Endress+Hauser
Page 53
Proline Promass F 300
T
a
T
m
B
A

Medium temperature range

Process

Standard version –50 to +150 °C (–58 to +302 °F) Order code for "Measuring tube
mat., wetted surface", option HA, SA, SB, SC
Extended temperature version –50 to +240 °C (–58 to +464 °F) Order code for "Measuring tube
mat., wetted surface", option SD, SE, SF, TH
High-temperature version –50 to +350 °C (–58 to +662 °F) For nominal diameters DN 25 (1"),
DN 50 (2") and DN 80 (3") Order code for "Measuring tube mat., wetted surface", option TT, TU
Low-temperature version –196 to +150 °C (–320 to +302 °F)
NOTICE
Material fatigue due to excessive temperature difference!
Maximum temperature
difference of media used: 300 K
Order code for "Measuring tube mat., wetted surface", option LA
Dependency of ambient temperature on medium temperature
 22 Exemplary representation, values in the table below.
TaAmbient temperature TmMedium temperature A Maximum permitted medium temperature Tm at T
require a reduced ambient temperature T
B Maximum permitted ambient temperature Ta for the maximum specified medium temperature Tm of the
sensor
a
= 60 °C (140 °F); higher medium temperatures T
a max
A0031121
m
Values for devices used in the hazardous area: Separate Ex documentation (XA) for the device →  121.
Endress+Hauser 53
Page 54
Proline Promass F 300
PN100
PN63
PN40
PN16
0
10
20
30
40
50
60
70
80
90
100
0
200
400
600
800
[bar][psi]
-320
64080 160
240
320 400 480 560
[°C]
[°F]
35030025020015010050
0
-200
Not insulated Insulated
A B A B
Version T
Standard version 60 °C (140 °F) 150 °C (302 °F) 60 °C (140 °F) 110 °C (230 °F) 55 °C (131 °F) 150 °C
Extended temperature version
High-temperature version
a
60 °C (140 °F) 170 °C (338 °F) 55 °C (131 °F) 240 °C
60 °C (140 °F) 350 °C (662 °F) 60 °C (140 °F) 350 °C (662 °F)
T
m
T
a
T
m
(464 °F)
T
a
60 °C (140 °F) 110 °C (230 °F) 50 °C (122 °F) 240 °C
T
m
T
a
T
m
(302 °F)
(464 °F)

Density

Pressure-temperature curves

0 to 5 000 kg/m3 (0 to 312 lb/cf)
The following pressure/temperature diagrams apply to all pressure-bearing parts of the device and not just the process connection. The diagrams show the maximum permissible medium pressure depending on the specific medium temperature.
• Pressure-temperature curves with temperature range +151 to +240 °C (+304 to +464 °F) exclusively for extended temperature version of measuring devices.
• Pressure-temperature curves with temperature range +241 to +350 °C (+466 to +662 °F) exclusively for high temperature version of measuring devices.
• Pressure-temperature curves with temperature range –196 to +150 °C (–320 to +302 °F) exclusively for low temperature version of measuring devices.
Flange according to EN 1092-1 (DIN 2501)
 23 With flange material 1.4404 (F316/F316L), Alloy C22
54 Endress+Hauser
A0032852-EN
Page 55
Proline Promass F 300
Class 600
Class 300
Class 150
0
10
20
30
40
50
60
70
80
90
100
0
200
400
600
800
[bar][psi]
-320
64080 160
240
320 400 480 560
[°C]
[°F]
35030025020015010050
0
-200
Class 600
Class 300
Class 150
-80
6400 80 160
240
320 400 480 560
0
10
20
30
40
50
60
70
80
90
100
0
200
400
600
800
[°C]
[°F]
[bar][psi]
35030025020015010050
0
-50
Flange according to ASME B16.5
A0032853-EN
 24 With flange material 1.4404 (F316/F316L)
A0028780-EN
 25 With flange material Alloy C22
Endress+Hauser 55
Page 56
Flange JIS B2220
63K
40K
20K
10K
0
10
20
30
40
50
60
70
0
200
400
600
800
[bar][psi]
-320
64080 160
240
320 400 480 560
[°C]
[°F]
35030025020015010050
0
-200
DN8…40
≥ DN50
-80
0 80 160
240
320 400 480
0
10
20
30
40
0
[°C]
[°F]
[bar][psi]
25020015010050
0
-50
50
PN40
-80
0 80 160
240
320 400 480
0
10
20
30
40
0
[°C]
[°F]
[bar][psi]
25020015010050
0
-50
50
 26 With flange material 1.4404 (F316/F316L), Alloy C22
Flange DIN 11864-2 Form A
Proline Promass F 300
A0032851-EN
 27 With connection material 1.4404 (316/316L)
Lap joint flange according to EN 1092-1 (DIN 2501)
 28 With flange material 1.4301 (F304); wetted parts Alloy C22
56 Endress+Hauser
A0028782-EN
A0028784-EN
Page 57
Proline Promass F 300
Class 150
Class 600
Class 300
-80
0 80 160
240
320 400 480
0
10
20
30
40
50
60
70
80
90
100
0
200
400
600
800
[°C]
[°F]
[bar][psi]
25020015010050
0
-50
20K
-80
0 80 160
240
320 400 480
0
10
20
30
0
200
400
[°C]
[°F]
[bar]
[psi]
25020015010050
0
-50
≥ DN50
DN8…40
-80
0 80 160
240
320 400 480
0
10
20
30
40
0
[°C]
[°F]
[bar][psi]
25020015010050
0
-50
50
Lap joint flange according to ASME B16.5
A0028785-EN
 29 With flange material 1.4301 (F304); wetted parts Alloy C22
Lap joint flange JIS B2220
A0028786-EN
 30 With flange material 1.4301 (F304); wetted parts Alloy C22
Thread DIN 11851
Endress+Hauser 57
 31 With connection material 1.4404 (316/316L)
DIN 11851 allows for applications up to +140 °C (+284 °F) if suitable sealing materials are used. Please take this into account when selecting seals and counterparts, as these components can limit the pressure and temperature range.
A0028794-EN
Page 58
Thread DIN 11864-1 Form A
DN8…40
≥ DN50
-80
0 80 160
240
320 400 480
0
10
20
30
40
0
[°C]
[°F]
[bar][psi]
25020015010050
0
-50
50
-80
0 80 160
240
320 400 480
0
10
20
0
[°C]
[°F]
[bar][psi]
25020015010050
0
-50
-80
0 80 160
240
320 400 480
0
10
20
0
[°C]
[°F]
[bar][psi]
25020015010050
0
-50
 32 With connection material 1.4404 (316/316L)
Thread ISO 2853
Proline Promass F 300
A0028798-EN
A0028799-EN
 33 With connection material 1.4404 (316/316L)
Thread SMS 1145
A0028800-EN
 34 With connection material 1.4404 (316/316L)
SMS 1145 allows for applications up to 16 bar (232 psi) if suitable sealing materials are used. Please take this into account when selecting seals and counterparts, as these components can limit the pressure and temperature range.
58 Endress+Hauser
Page 59
Proline Promass F 300
60
70
80
90
100
800
[bar][psi]
-80
0 80 160
240
320 400 480
[°C]
[°F]
25020015010050
0
-50
-80
0 80 160
240
320 400 480
0
10
20
0
[°C]
[°F]
[bar][psi]
25020015010050
0
-50
VCO
A0028801-EN
 35 With connection material 1.4404 (316/316L)
Tri-Clamp
The clamp connections are suitable up to a maximum pressure of 16 bar (232 psi). Please observe the operating limits of the clamp and seal used as they can be over 16 bar (232 psi). The clamp and seal are not included in the scope of supply.

Sensor housing

For standard versions with the temperature range –50 to +150 °C (–58 to +302 °F), the sensor housing is filled with dry nitrogen gas and protects the electronics and mechanics inside.
For all other temperature versions the sensor housing is filled with dry inert gas.
If a measuring tube fails (e.g. due to process characteristics like corrosive or abrasive fluids), the fluid will initially be contained by the sensor housing.
In the event of a tube failure, the pressure level inside the sensor housing will rise according to the operating process pressure. If the user judges that the sensor housing pressure rating/burst pressure does not provide an adequate safety margin, the device can be fitted with a rupture disk. This prevents excessively high pressure from forming inside the sensor housing. Therefore, the use of a rupture disk is strongly recommended in applications involving high gas pressures, and particularly in applications in which the process pressure is greater than 2/3 of the sensor housing burst pressure.
If there is a need to drain the leaking medium into a discharge device, the sensor should be fitted with a rupture disk. Connect the discharge to the additional threaded connection →  82.
If the sensor is to be purged with gas (gas detection), it should be equipped with purge connections.
Do not open the purge connections unless the containment can be filled immediately with a dry, inert gas. Use only low pressure to purge. Maximum pressure: 5 bar (72.5 psi).
Endress+Hauser 59
Sensor housing nominal pressure rating and burst pressure
The following sensor housing nominal pressure ratings/burst pressures are only valid for standard devices and/or devices equipped with closed purge connections (not opened/as delivered).
A0032216-EN
Page 60
Proline Promass F 300
If a device fitted with purge connections (order code for "Sensor option", option CH "Purge connection") is connected to the purge system, the maximum nominal pressure is determined by the purge system itself or by the device, depending on which component has the lower nominal pressure classification.
If the device is fitted with a rupture disk (order code for "Sensor option", option CA "Rupture disk"), the rupture disk trigger pressure is decisive for the maximum nominal pressure .
The sensor housing burst pressure refers to a typical internal pressure which is reached prior to mechanical failure of the sensor housing and which was determined during type testing. The corresponding type test declaration can be ordered with the device (order code for "Additional approval", option LN "Sensor housing burst pressure, type test").

Rupture disk

DN Sensor housing nominal
pressure
(designed with a safety factor
≥ 4)
[mm] [in] [bar] [psi] [bar] [psi]
8 ³⁄₈ 40 580 255 3 698
15 ½ 40 580 200 2 900
25 1 40 580 280 4 060
40 40 580 180 2 610
50 2 40 580 195 2 828
80 3 25 362 105 1 522
100 4 16 232 85 1 232
150 6 16 232 80 1 160
250 10 10 145 57 826
Sensor housing burst pressure
For information on the dimensions: see the "Mechanical construction" section
To increase the level of safety, a device version with a rupture disk with a trigger pressure of 10 to 15 bar (145 to 217.5 psi)can be used (order code for "Sensor option", option CA "rupture disk").
The use of rupture disks cannot be combined with the separately available heating jacket.
Information on the dimensions of the rupture disk: →  82

Flow limit

Select the nominal diameter by optimizing between the required flow range and permissible pressure loss.
For an overview of the full scale values for the measuring range, see the "Measuring range" section →  11
• The minimum recommended full scale value is approx. 1/20 of the maximum full scale value
• In most applications, 20 to 50 % of the maximum full scale value can be considered ideal
• A low full scale value must be selected for abrasive media (such as liquids with entrained solids): flow velocity < 1 m/s (< 3 ft/s).
• For gas measurement the following rules apply: – The flow velocity in the measuring tubes should not exceed half the sound velocity (0.5 Mach). – The maximum mass flow depends on the density of the gas: formula →  11
To calculate the flow limit, use the Applicator sizing tool →  119

Pressure loss

To calculate the pressure loss, use the Applicator sizing tool →  119
Promass F with reduced pressure loss: order code for "Sensor option", option CE "Reduced pressure loss"

System pressure

It is important that cavitation does not occur, or that gases entrained in the liquids do not outgas. This is prevented by means of a sufficiently high system pressure.
60 Endress+Hauser
Page 61
Proline Promass F 300
For this reason, the following mounting locations are recommended:
• At the lowest point in a vertical pipe
• Downstream from pumps (no danger of vacuum)
A0028777

Thermal insulation

In the case of some fluids, it is important to keep the heat radiated from the sensor to the transmitter to a low level. A wide range of materials can be used for the required insulation.
The following device versions are recommended for versions with thermal insulation:
• Version with extended neck for insulation: Order code for "Sensor option", option CG with an extended neck length of 105 mm (4.13 in).
• Extended temperature version: Order code for "Measuring tube material", option SD, SE, SF or TH with an extended neck length of 105 mm (4.13 in).
• High-temperature version: Order code for "Measuring tube material", option TT or TU with an extended neck length of 142 mm (5.59 in).
NOTICE
Electronics overheating on account of thermal insulation!
Recommended orientation: horizontal orientation, transmitter housing pointing downwards.
Do not insulate the transmitter housing .
Maximum permissible temperature at the lower end of the transmitter housing: 80 °C (176 °F)
Thermal insulation with extended neck free: We recommend that you do not insulate the
extended neck in order to ensure optimum dissipation of heat.
A0034391
 36 Thermal insulation with extended neck free
Low-temperature version: It is generally not necessary to insulate the transmitter housing . If insulation is provided, the rules that apply are the same as those for thermal insulation.

Heating

Some fluids require suitable measures to avoid loss of heat at the sensor.
Heating options
• Electrical heating, e.g. with electric band heaters
• Via pipes carrying hot water or steam
• Via heating jackets
Heating jackets for the sensors can be ordered as accessories from Endress+Hauser .→  118
Endress+Hauser 61
Page 62
Proline Promass F 300
NOTICE
Danger of overheating when heating
Ensure that the temperature at the lower end of the transmitter housing does not exceed
80 °C (176 °F). Ensure that sufficient convection takes place at the transmitter neck.
Ensure that a sufficiently large area of the transmitted neck remains exposed. The uncovered
part serves as a radiator and protects the electronics from overheating and excessive cooling. If using in potentially explosive atmospheres, observe the information in the device-specific Ex
documentation. For detailed information on the temperature tables, see the separate document entitled "Safety Instructions" (XA) for the device.

Vibrations

The high oscillation frequency of the measuring tubes ensures that the correct operation of the measuring system is not influenced by plant vibrations.
62 Endress+Hauser
Page 63
Proline Promass F 300

Custody transfer measurement

The measuring device is optionally tested in accordance with OIML R117/R81 and has an EU type evaluation certificate which authorizes the use in EU type-examination certificates according to Measuring Instruments Directive 2014/32/EU for service subject to legal metrological control ("custody transfer") for liquids other than water and cryogenic liquids (Annex VII).
The permitted fluid temperature in these applications is –200 to +90 °C (–328 to +194 °F).
The measuring device is optionally tested in accordance with OIML R137 and has an EU type­examination certificate according to Measuring Instruments Directive 2014/32/EU for service as a gas meter subject to legal metrological control ("custody transfer") (Annex IV). The permitted fluid temperature in these applications is –25 to +55 °C (–13 to +131 °F).
The device is used with a legally controlled totalizer on the local display and optionally with legally controlled outputs.
Measuring devices subject to legal metrological control totalize in both directions, i.e. all the outputs consider flow components in the positive (forward) and negative (reverse) flow direction.
Generally a measuring device subject to legal metrological control is secured against tampering by seals on the transmitter or sensor. These seals may normally only be opened by a representative of the competent authority for legal metrology controls.
After putting the device into circulation or after sealing the device, operation is only possible to a limited extent.
Detailed ordering information is available from your local Endress+Hauser sales center for national approvals, which are based on the OIML certificates, for applications with liquids other than water, cryogenic liquids or gases.
Endress+Hauser 63
Page 64

Mechanical construction

D
L
E
F
A
CB
K
G
H
I
M

Dimensions in SI units Compact version

Proline Promass F 300
A0033786
Order code for "Housing", option A "Aluminum, coated"
DN A
1)
B
C D E
2) 3)F 2) 3)
4)
G
H I
4)
K L M
1)
[mm] [mm] [mm] [mm] [mm] [mm] [mm] [mm] [mm] [mm] [mm] [mm] [mm]
8 169 68 101 75 259.5 334.5 200 59 141 5.35
15 169 68 101 75 259.5 334.5 200 59 141 8.30
25 169 68 101 75 259.5 334.5 200 59 141 12.0
40 169 68 101 105 264.5 369.5 200 59 141 17.6
50 169 68 101 141 274.5 415.5 200 59 141 26.0
80 169 68 101 200 294.5 494.5 200 59 141 40.5
100 169 68 101 254 312.5 566.5 200 59 141 51.2
150 169 68 101 378 333.5 711.5 200 59 141 68.9
250 169 68 101 548 377.5 925.5 200 59 141 102.3
5)
70
5)
70
5)
70
5)
79
5)
99
5)
139
5)
176
5)
218
5)
305
1) Depending on the cable gland used: values up to + 30 mm
2) With order code for "Sensor option", option CG or order code for "Measuring tube material", option SD, SE, SF, TH, LA: values +70 mm
3) With order code for "Measuring tube material", option TT, TU: values +104 mm
4) For version without local display: values - 30 mm
5) Depends on the process connection in question
Order code for "Housing", option A "Aluminum, coated"; Ex d
1)
DN A
1)
B
C D E
[mm] [mm] [mm] [mm] [mm] [mm] [mm] [mm] [mm] [mm] [mm] [mm] [mm]
8 188 85 103 75 290 365 206 58 148 5.35
15 188 85 103 75 290 365 206 58 148 8.30
25 188 85 103 75 290 365 206 58 148 12.0
40 188 85 103 105 294.5 399.5 206 58 148 17.6
50 188 85 103 141 304.5 445.5 206 58 148 26.0
80 188 85 103 200 324.5 524.5 206 58 148 40.5
100 188 85 103 254 342.5 596.5 206 58 148 51.2
2) 3)
F G
4)
H I
4)
K L M
5)
70
5)
70
5)
70
5)
79
5)
99
5)
139
5)
176
64 Endress+Hauser
Page 65
Proline Promass F 300
DN A
1)
B
C D E
2) 3)
F G
4)
H I
4)
K L M
1)
[mm] [mm] [mm] [mm] [mm] [mm] [mm] [mm] [mm] [mm] [mm] [mm] [mm]
150 188 85 103 378 363.5 741.5 206 58 148 68.9
250 188 85 103 548 407.5 955.5 206 58 148 102.3
5)
218
5)
305
1) Depending on the cable gland used: values up to + 30 mm
2) With order code for "Sensor option", option CG or order code for "Measuring tube material", option SD, SE, SF, TH, LA: values +70 mm
3) With order code for "Measuring tube material", option TT, TU: values +104 mm
4) For version without local display: values - 38 mm
5) Depends on the process connection in question
Order code for "Housing", option B "Stainless, hygienic"
1)
DN A
1)
B
C D E
[mm] [mm] [mm] [mm] [mm] [mm] [mm] [mm] [mm] [mm] [mm] [mm] [mm]
8 183 73 110 75 259.5 334.5 207 65 142 5.35
15 183 73 110 75 259.5 334.5 207 65 142 8.30
25 183 73 110 75 259.5 334.5 207 65 142 12.0
40 183 73 110 105 264.5 369.5 207 65 142 17.6
50 183 73 110 141 274.5 415.5 207 65 142 26.0
80 183 73 110 200 294.5 494.5 207 65 142 40.5
100 183 73 110 254 312.5 566.5 207 65 142 51.2
150 183 73 110 378 333.5 711.5 207 65 142 68.9
250 183 73 110 548 377.5 925.5 207 65 142 102.3
2) 3)
F G
4)
H I
4)
K L M
5)
70
5)
70
5)
70
5)
79
5)
99
5)
139
5)
176
5)
218
5)
305
1) Depending on the cable gland used: values up to + 30 mm
2) With order code for "Sensor option", option CG or order code for "Measuring tube material", option SD, SE, SF, TH, LA: values +70 mm
3) With order code for "Measuring tube material", option TT, TU: values +104 mm
4) For version without local display: values - 13 mm
5) Depends on the process connection in question
Order code for "Housing", option L "Cast, stainless"
1)
DN A
1)
B
C D E
[mm] [mm] [mm] [mm] [mm] [mm] [mm] [mm] [mm] [mm] [mm] [mm] [mm]
8 186 85 101 75 290 365 221 63 158 5.35
15 186 85 101 75 290 365 221 63 158 8.30
25 186 85 101 75 290 365 221 63 158 12.0
40 186 85 101 105 294.5 399.5 221 63 158 17.6
50 186 85 101 141 304.5 445.5 221 63 158 26.0
80 186 85 101 200 324.5 524.5 221 63 158 40.5
100 186 85 101 254 342.5 596.5 221 63 158 51.2
150 186 85 101 378 363.5 741.5 221 63 158 68.9
250 186 85 101 548 407.5 955.5 221 63 158 102.3
1) Depending on the cable gland used: values up to + 30 mm
2) With order code for "Sensor option", option CG or order code for "Measuring tube material", option SD, SE, SF, TH, LA: values +70 mm
3) With order code for "Measuring tube material", option TT, TU: values +104 mm
4) Depends on the process connection in question
2) 3)
F G H I K L M
4)
70
4)
70
4)
70
4)
79
4)
99
4)
139
4)
176
4)
218
4)
305
Endress+Hauser 65
Page 66
Flange connections
C
D
A
B
E
L
Fixed flange EN 1092-1, ASME B16.5, JIS B2220
Length tolerance for dimension L in mm:
• DN ≤ 100: +1.5 / –2.0
• DN ≥ 125: +3.5
Proline Promass F 300
A0015621
Flange according to EN 1092-1 (DIN 2501): PN16
1.4404 (F316/F316L): order code for "Process connection", option D1S Alloy C22: order code for "Process connection", option D1C
Flange with groove according to EN 1092-1 Form D (DIN 2512N): PN16
1.4404 (F316/F316L): order code for "Process connection", option D5S Alloy C22: order code for "Process connection", option D5C
DN
[mm]
100 220 180 8 × ⌀18 20 107.1 1 127/1 400
150 285 240 8 × ⌀22 22 159.3 1 330/1 700
250 405 355 12 × ⌀26 26 260.4 1 775
Surface roughness (flange): EN 1092-1 Form B1 (DIN 2526 Form C), Ra 3.2 to 12.5 µm
1) Installation length in accordance with NAMUR recommendation NE 132 optionally available (order code for "Process connection", option D1N or D5N (with groove))
Flange according to EN 1092-1 (DIN 2501): PN16 with reduction in nominal diameter
1.4404 (F316/F316L
DN
[mm]
100 80 DHS 220 180 8 × ⌀ 18 20 107.1 874
150 100 DJS 285 240 8 × ⌀ 22 22 159.3 1 167
200 150 DLS 340 295 12 × ⌀ 22 24 206.5 1 461
Surface roughness (flange): EN 1092-1 Form B1 (DIN 2526 Form C), Ra 3.2 to 12.5 µm
A
[mm]
reduction
to DN [mm]
B
[mm]
Order code for
"Process
connection",
option
C
[mm]
A
[mm]B[mm]
D
[mm]
C
[mm]
E
[mm]
D
[mm]E[mm]L[mm]
L
[mm]
1)
1)
66 Endress+Hauser
Page 67
Proline Promass F 300
Flange according to EN 1092-1 (DIN 2501): PN 40
1.4404 (F316/F316L): order code for "Process connection", option D2S Alloy C22: order code for "Process connection", option D2C
Flange with groove according to EN 1092-1 Form D (DIN 2512N): PN 40
1.4404 (F316/F316L): order code for "Process connection", option D6S Alloy C22: order code for "Process connection", option D6C
DN
[mm]
1)
8
A
[mm]
B
[mm]
C
[mm]
D
[mm]
E
[mm]
95 65 4 × ⌀14 16 17.3 370/510
15 95 65 4 × ⌀14 16 17.3 404/510
25 115 85 4 × ⌀14 18 28.5 440/600
40 150 110 4 × ⌀18 18 43.1 550
50 165 125 4 × ⌀18 20 54.5 715/715
80 200 160 8 × ⌀18 24 82.5 840/915
100 235 190 8 × ⌀22 24 107.1 1 127
150 300 250 8 × ⌀26 28 159.3 1 370
250 450 385 12 × ⌀33 38 258.8 1 845
Surface roughness (flange): EN 1092-1 Form B1 (DIN 2526 Form C), Ra 3.2 to 12.5 µm
L
[mm]
2)
2)
2)
2)
2)
1) DN 8 with DN 15 flanges as standard
2) Installation length in accordance with NAMUR recommendation NE 132 optionally available (order code for "Process connection", option D2N or D6N (with groove))
Flange according to EN 1092-1 (DIN 2501): PN 40 (with DN 25 flanges)
1.4404 (F316/F316L): order code for "Process connection", option R2S
DN
[mm]
A
[mm]
B
[mm]
C
[mm]
D
[mm]
E
[mm]
L
[mm]
8 115 85 4 × ⌀14 18 28.5 440
15 115 85 4 × ⌀14 18 28.5 440
Surface roughness (flange): EN 1092-1 Form B1 (DIN 2526 Form C), Ra 3.2 to 12.5 µm
Flange according to EN 1092-1 (DIN 2501): PN 40 with reduction in nominal diameter
1.4404 (F316/F316L)
DN
[mm]
reduction
to DN [mm]
Order code for
"Process
connection",
A
[mm]B[mm]
C
[mm]
D
[mm]E[mm]L[mm]
option
50 40 DFS 165 125 4 × ⌀ 18 20 54.5 555
80 50 DGS 200 160 8 × ⌀ 18 24 82.5 840
100 80 DIS 235 190 8 × ⌀ 22 24 107.1 874
150 100 DKS 300 250 8 × ⌀ 26 28 159.3 1 167
200 150 DMS 375 320 12 × ⌀ 30 34 206.5 1 461
Surface roughness (flange): EN 1092-1 Form B1 (DIN 2526 Form C), Ra 3.2 to 12.5 µm
Endress+Hauser 67
Page 68
Proline Promass F 300
Flange according to EN 1092-1 (DIN 2501): PN 63
1.4404 (F316/F316L): order code for "Process connection", option D3S Alloy C22: order code for "Process connection", option D3C
Flange with groove according to EN 1092-1 Form D (DIN 2512N): PN 63
1.4404 (F316/F316L): order code for "Process connection", option D7S Alloy C22: order code for "Process connection", option D7C
DN
[mm]
50 180 135 4 × ⌀22 26 54.5 724
80 215 170 8 × ⌀22 28 81.7 875
100 250 200 8 × ⌀26 30 106.3 1 127
150 345 280 8 × ⌀33 36 157.1 1 410
250 470 400 12 × ⌀36 46 255.4 1 885
Surface roughness (flange): EN 1092-1 Form B1 (DIN 2526 Form C), Ra 3.2 to 12.5 µm EN 1092-1 Form B2 (DIN 2526 Form E), Ra
0.8 to 3.2 µm
Flange according to EN 1092-1 (DIN 2501): PN 100
1.4404 (F316/F316L): order code for "Process connection", option D4S Alloy C22: order code for "Process connection", option D4C
Flange with groove according to EN 1092-1 Form D (DIN 2512N): PN 100
1.4404 (F316/F316L): order code for "Process connection", option D8S Alloy C22: order code for "Process connection", option D8C
DN
[mm]
1)
8
15 105 75 4 × ⌀14 20 17.3 420
25 140 100 4 × ⌀18 24 28.5 470
40 170 125 4 × ⌀22 26 42.5 590
50 195 145 4 × ⌀26 28 53.9 740
80 230 180 8 × ⌀26 32 80.9 885
100 265 210 8 × ⌀30 36 104.3 1 127
150 355 290 12 × ⌀33 44 154.0 1 450
Surface roughness (flange): EN 1092-1 Form B2 (DIN 2526 Form E), Ra 0.8 to 3.2 µm
A
[mm]
A
[mm]
105 75 4 × ⌀14 20 17.3 400
B
[mm]
B
[mm]
C
[mm]
C
[mm]
D
[mm]
D
[mm]
E
[mm]
E
[mm]
[mm]
[mm]
L
L
1) DN 8 with DN 15 flanges as standard
Flange according to EN 1092-1 (DIN 2501): PN 100 Alloy C22: order code for "Process connection", option D4C
Flange with groove according to EN 1092-1 Form D (DIN 2512N): PN 100 Alloy C22: order code for "Process connection", option D8C
DN
[mm]
250 505 430 12 × ⌀39 60 248.0 1 949
Surface roughness (flange): EN 1092-1 Form B2 (DIN 2526 Form E), Ra 0.8 to 3.2 µm
A
[mm]
B
[mm]
C
[mm]
D
[mm]
E
[mm]
L
[mm]
68 Endress+Hauser
Page 69
Proline Promass F 300
Flange according to ASME B16.5: Class 150
1.4404 (F316/F316L): order code for "Process connection", option AAS Alloy C22: order code for "Process connection", option AAC
DN
[mm]
1)
8
A
[mm]
B
[mm]
C
[mm]
D
[mm]
E
[mm]
90 60.3 4 × ⌀15.7 11.2 15.7 370
15 90 60.3 4 × ⌀15.7 11.2 15.7 404
25 110 79.4 4 × ⌀15.7 14.2 26.7 440
40 125 98.4 4 × ⌀15.7 17.5 40.9 550
50 150 120.7 4 × ⌀19.1 19.1 52.6 715
80 190 152.4 4 × ⌀19.1 23.9 78.0 840
100 230 190.5 8 × ⌀19.1 23.9 102.4 1 127
150 280 241.3 8 × ⌀22.4 25.4 154.2 1 398
250 405 362 12 × ⌀25.4 30.2 254.5 1 832
Surface roughness (flange): Ra 3.2 to 6.3 µm
1) DN 8 with DN 15 flanges as standard
L
[mm]
Flange according to ASME B16.5: Class 150 with reduction in nominal diameter
1.4404 (F316/F316L)
DN
[mm]
reduction
to DN [mm]
Order code for
"Process
connection",
A
[mm]B[mm]
C
[mm]
D
[mm]E[mm]L[mm]
option
50 40 AHS 150 120.7 4 × ⌀ 19.1 19.1 52.6 550
80 50 AJS 190 152.4 4 × ⌀ 19.1 23.9 78.0 720
100 80 ALS 230 190.5 8 × ⌀ 19.1 23.9 102.4 874
150 100 ANS 280 241.3 8 × ⌀ 22.4 25.4 154.2 1 167
200 150 APS 345 298.5 8 × ⌀ 22.4 29 202.7 1 461
Surface roughness (flange): Ra 3.2 to 6.3 µm
Flange according to ASME B16.5: Class 300
1.4404 (F316/F316L): order code for "Process connection", option ABS Alloy C22: order code for "Process connection", option ABC
DN
[mm]
1)
8
A
[mm]
B
[mm]
C
[mm]
D
[mm]
E
[mm]
95 66.7 4 × ⌀15.7 14.2 15.7 370
[mm]
15 95 66.7 4 × ⌀15.7 14.2 15.7 404
25 125 88.9 4 × ⌀19.1 17.5 26.7 440
40 155 114.3 4 × ⌀22.3 20.6 40.9 550
50 165 127 8 × ⌀19.1 22.3 52.6 715
80 210 168.3 8 × ⌀22.3 28.4 78.0 840
100 255 200 8 × ⌀22.3 31.7 102.4 1 127
150 320 269.9 12 × ⌀22.3 36.5 154.2 1 417
250 445 387.4 16 × ⌀28.4 47.4 254.5 1 863
Surface roughness (flange): Ra 3.2 to 6.3 µm
L
1) DN 8 with DN 15 flanges as standard
Endress+Hauser 69
Page 70
Proline Promass F 300
Flange according to ASME B16.5: Class 300 with reduction in nominal diameter
1.4404 (F316/F316L)
DN
[mm]
50 40 AIS 165 127 8 × ⌀ 19.1 22.3 52.6 615
80 50 AKS 210 168.3 8 × ⌀ 22.3 28.4 78.0 732
100 80 AMS 255 200 8 × ⌀ 22.3 31.7 102.4 894
150 100 AOS 320 269.9 12 × ⌀ 22.3 36.5 154.2 1 187
200 150 AQS 380 330.2 12 × ⌀ 25.4 41.7 202.7 1 461
Surface roughness (flange): Ra 3.2 to 6.3 µm
Flange according to ASME B16.5: Class 600
1.4404 (F316/F316L): order code for "Process connection", option ACS Alloy C22: order code for "Process connection", option ACC
Surface roughness (flange): Ra 3.2 to 6.3 µm
reduction
to DN [mm]
DN
[mm]
1)
8
15 95 66.7 4 × ⌀15.7 20.6 13.9 420
25 125 88.9 4 × ⌀19.1 23.9 24.3 490
40 155 114.3 4 × ⌀22.3 28.7 38.1 600
50 165 127 8 × ⌀19.1 31.8 49.2 742
80 210 168.3 8 × ⌀22.3 38.2 73.7 900
100 275 215.9 8 × ⌀25.4 48.4 97.3 1 157
150 355 292.1 12 × ⌀28.4 47.8 154.2 1 467
250 510 431.8 16 × ⌀35.1 69.9 254.5 1 946
[mm]
Order code for
"Process
connection",
option
A
95 66.7 4 × ⌀15.7 20.6 13.9 400
B
[mm]
A
[mm]B[mm]
C
[mm]
C
[mm]
D
[mm]
D
[mm]E[mm]L[mm]
E
[mm]
[mm]
L
1) DN 8 with DN 15 flanges as standard
Flange JIS B2220: 10K
1.4404 (F316/F316L): order code for "Process connection", option NDS Alloy C22: order code for "Process connection", option NDC
DN
[mm]
50 155 120 4 × ⌀19 16 50 715
80 185 150 8 × ⌀19 18 80 832
100 210 175 8 × ⌀19 18 100 1 127
150 280 240 8 × ⌀23 22 150 1 354
250 400 355 12 × ⌀25 24 250 1 775
Surface roughness (flange): Ra 3.2 to 6.3 µm
A
[mm]
B
[mm]
C
[mm]
D
[mm]
E
[mm]
L
[mm]
70 Endress+Hauser
Page 71
Proline Promass F 300
Flange JIS B2220: 20K
1.4404 (F316/F316L): order code for "Process connection", option NES Alloy C22: order code for "Process connection", option NEC
DN
[mm]
1)
8
A
[mm]
B
[mm]
C
[mm]
D
[mm]
E
[mm]
95 70 4 × ⌀15 14 15 370
15 95 70 4 × ⌀15 14 15 404
25 125 90 4 × ⌀19 16 25 440
40 140 105 4 × ⌀19 18 40 550
50 155 120 8 × ⌀19 18 50 715
80 200 160 8 × ⌀23 22 80 832
100 225 185 8 × ⌀23 24 100 1 127
150 305 260 12 × ⌀25 28 150 1 386
250 430 380 12 × ⌀27 34 250 1 845
Surface roughness (flange): Ra 1.6 to 3.2 µm
1) DN 8 with DN 15 flanges as standard
L
[mm]
Flange JIS B2220: 40K
1.4404 (F316/F316L): order code for "Process connection", option NGS Alloy C22: order code for "Process connection", option NGC
DN
[mm]
1)
8
A
[mm]
B
[mm]
C
[mm]
D
[mm]
E
[mm]
115 80 4 × ⌀19 20 15 400
15 115 80 4 × ⌀19 20 15 425
25 130 95 4 × ⌀19 22 25 485
40 160 120 4 × ⌀23 24 38 600
50 165 130 8 × ⌀19 26 50 760
80 210 170 8 × ⌀23 32 75 890
100 250 205 8 × ⌀25 36 100 1 167
150 355 295 12 × ⌀33 44 150 1 498
Surface roughness (flange): Ra 1.6 to 3.2 µm
1) DN 8 with DN 15 flanges as standard
Flange JIS B2220: 63K
1.4404 (F316/F316L): order code for "Process connection", option NHS Alloy C22: order code for "Process connection", option NHC
DN
[mm]
1)
8
A
[mm]
B
[mm]
C
[mm]
D
[mm]
E
[mm]
120 85 4 × ⌀19 23 12 420
15 120 85 4 × ⌀19 23 12 440
25 140 100 4 × ⌀23 27 22 494
40 175 130 4 × ⌀25 32 35 620
50 185 145 8 × ⌀23 34 48 775
80 230 185 8 × ⌀25 40 73 915
100 270 220 8 × ⌀27 44 98 1 167
L
[mm]
L
[mm]
Endress+Hauser 71
Page 72
Proline Promass F 300
Flange JIS B2220: 63K
1.4404 (F316/F316L): order code for "Process connection", option NHS Alloy C22: order code for "Process connection", option NHC
DN
[mm]
150 365 305 12 × ⌀33 54 146 1 528
Surface roughness (flange): Ra 1.6 to 3.2 µm
1) DN 8 with DN 15 flanges as standard
A
[mm]
B
[mm]
C
[mm]
D
[mm]
E
[mm]
L
[mm]
72 Endress+Hauser
Page 73
Proline Promass F 300
C
D
A
B
E
X X
L
Fixed flange DIN 11864-2
A0015627
 37 Detail X: Asymmetrical process connection; the part shown in blue is provided by the supplier.
Length tolerance for dimension L in mm: +1.5 / –2.0
Flange DIN11864-2 Form A, for pipe according to DIN11866 series A, flat with notch
1.4404 (316/316L) Order code for "Process connection", option KCS
DN
[mm]
8 54 37 4 × ⌀9 10 10 387
15 59 42 4 × ⌀9 10 16 418
25 70 53 4 × ⌀9 10 26 454
40 82 65 4 × ⌀9 10 38 560
50 94 77 4 × ⌀9 10 50 720
80 133 112 8 × ⌀11 12 81 900
100 159 137 8 × ⌀11 14 100 1 127
3A-version available: order code for "Additional approval", option LP in conjunction with Ra ≤ 0.8 µm: order code for "Measuring tube material", option SB, SE or Ra ≤ 0.4 µm: order code for "Measuring tube material", option SC, SF
A
[mm]
B
[mm]
C
[mm]
D
[mm]
E
[mm]
L
[mm]
Endress+Hauser 73
Page 74
Proline Promass F 300
L
C
D
A
B
F
E
Lap joint flange EN 1092-1, ASME B16.5, JIS B2220
Length tolerance for dimension L in mm: +1.5 / –2.0
Lap joint flange according to EN 1092-1 Form D: PN 40
1.4301 (F304), wetted parts Alloy C22 Order code for "Process connection", option DAC
DN
[mm]A[mm]B[mm]
2)
8
95 65 4 × ⌀ 14 14.5 45 17.3 370 0
15 95 65 4 × ⌀ 14 14.5 45 17.3 404 0
25 115 85 4 × ⌀ 14 16.5 68 28.5 444 +4
40 150 110 4 × ⌀ 18 21 88 43.1 560 +10
50 165 125 4 × ⌀ 18 23 102 54.5 719 +4
80 200 160 8 × ⌀ 18 29 138 82.5 848 +8
100 235 190 8 × ⌀ 22 34 162 107.1 1 131 +4
Surface roughness (flange): Ra 3.2 to 12.5 µm
C
[mm]
D
[mm]E[mm]F[mm]L[mm]
A0022221
L
diff
[mm]
1)
1) Difference to installation length of the welding neck flange (order code for "Process connection", option D2C)
2) DN 8 with DN 15 flanges as standard
Lap joint flange according to ASME B16.5: Class 150
1.4301 (F304), wetted parts Alloy C22 Order code for "Process connection", option ADC
DN
[mm]A[mm]B[mm]
2)
8
90 60.3 4 × ⌀ 15.7 15 35.1 15.7 370 0
C
[mm]
D
[mm]E[mm]F[mm]L[mm]
L
diff
[mm]
1)
15 90 60.3 4 × ⌀ 15.7 15 35.1 15.7 404 0
25 110 79.4 4 × ⌀ 15.7 16 50.8 26.7 440 0
40 125 98.4 4 × ⌀ 15.7 15.9 73.2 40.9 550 0
50 150 120.7 4 × ⌀ 19.1 19 91.9 52.6 715 0
80 190 152.4 4 × ⌀ 19.1 22.3 127.0 78.0 840 0
74 Endress+Hauser
Page 75
Proline Promass F 300
Lap joint flange according to ASME B16.5: Class 150
1.4301 (F304), wetted parts Alloy C22 Order code for "Process connection", option ADC
DN
[mm]A[mm]
B
[mm]
C
[mm]
D
[mm]E[mm]F[mm]L[mm]
L
diff
[mm]
100 230 190.5 8 × ⌀ 19.1 26 157.2 102.4 1 127 0
Surface roughness (flange): Ra 3.2 to 12.5 µm
1) Difference to installation length of the welding neck flange (order code for "Process connection", option AAC)
2) DN 8 with DN 15 flanges as standard
Lap joint flange according to ASME B16.5: Class 300
1.4301 (F304), wetted parts Alloy C22 Order code for "Process connection", option AEC
DN
[mm]A[mm]
2)
8
95 66.7 4 × ⌀ 15.7 16.5 35.1 15.7 376 +6
B
[mm]
C
[mm]
D
[mm]E[mm]F[mm]L[mm]
L
diff
[mm]
15 95 66.7 4 × ⌀ 15.7 16.5 35.1 15.7 406 +2
25 125 88.9 4 × ⌀ 19.1 21.0 50.8 26.7 450 +10
40 155 114.3 4 × ⌀ 22.3 23.0 73.2 40.9 564 +14
50 165 127 8 × ⌀ 19.1 25.5 91.9 52.6 717 +2
80 210 168.3 8 × ⌀ 22.3 31.0 127.0 78.0 852.6 +12.6
100 255 200 8 × ⌀ 22.3 32.0 157.2 102.4 1 139 +12
Surface roughness (flange): Ra 3.2 to 12.5 µm
1)
1)
1) Difference to installation length of the welding neck flange (order code for "Process connection", option ABC)
2) DN 8 with DN 15 flanges as standard
Lap joint flange according to ASME B16.5: Class 600
1.4301 (F304), wetted parts Alloy C22 Order code for "Process connection", option AFC
DN
[mm]A[mm]
2)
8
95 66.7 4 × ⌀ 15.7 17.0 35.1 13.9 400 0
B
[mm]
C
[mm]
D
[mm]E[mm]F[mm]L[mm]
L
diff
[mm]
15 95 66.7 4 × ⌀ 15.7 17.0 35.1 13.9 420 0
25 125 88.9 4 × ⌀ 19.1 21.5 50.8 24.3 490 0
40 155 114.3 4 × ⌀ 22.3 25.0 73.2 38.1 600 0
50 165 127 8 × ⌀ 19.1 28.0 91.9 49.2 742 0
80 210 168.3 8 × ⌀ 22.3 35.0 127.0 73.7 900 0
100 275 215.9 8 × ⌀ 25.4 44.0 157.2 97.3 1 167 +10
Surface roughness (flange): Ra 3.2 to 12.5 µm
1) Difference to installation length of the welding neck flange (order code for "Process connection", option ACC)
2) DN 8 with DN 15 flanges as standard
1)
Endress+Hauser 75
Page 76
Proline Promass F 300
Lap joint flange JIS B2220: 20K
1.4301 (F304), wetted parts Alloy C22 Order code for "Process connection", option NIC
DN
[mm]
2)
8
15 95 70 4 × ⌀ 15 14 51 15 404 0
25 125 90 4 × ⌀ 19 18.5 67 25 440 0
40 140 105 4 × ⌀ 19 18.5 81 40 550 0
50 155 120 8 × ⌀ 19 23 96 50 715 0
80 200 160 8 × ⌀ 23 29 132 80 844 +12
100 225 185 8 × ⌀ 23 29 160 100 1 127 0
Surface roughness (flange): Ra 3.2 to 12.5 µm
1) Difference to installation length of the welding neck flange (order code for "Process connection", option NEC)
2) DN 8 with DN 15 flanges as standard
A
[mm]
B
[mm]
C
[mm]
D
[mm]
E
[mm]
F
[mm]
L
[mm]
95 70 4 × ⌀ 15 14 51 15 370 0
L
diff
[mm]
1)
76 Endress+Hauser
Page 77
Proline Promass F 300
L
A
B
Clamp connections
Tri-Clamp
Length tolerance for dimension L in mm: +1.5 / –2.0
Tri-Clamp (½"), for pipe according to DIN 11866 series C
1.4404 (316/316L) Order code for "Process connection", option FDW
DN
[mm]
8 ½ 25.0 9.5 367
15 ½ 25.0 9.5 398
3-A version available: order code for "Additional approval", option LP in conjunction with Ra ≤ 0.8 µm: order code for "Measuring tube material", option SB, SE or Ra ≤ 0.4 µm: order code for "Measuring tube material", option SC, SF
Clamp
[in]
A
[mm]
B
[mm]
A0015625
L
[mm]
Tri-Clamp ( ≥ 1"), for pipe according to DIN 11866 series C
1.4404 (316/316L) Order code for "Process connection", option FTS
DN
[mm]
8 1 50.4 22.1 367
15 1 50.4 22.1 398
25 1 50.4 22.1 434
40 50.4 34.8 560
50 2 63.9 47.5 720
80 3 90.9 72.9 900
100 4 118.9 97.4 1 127
3-A version available: order code for "Additional approval", option LP in conjunction with Ra ≤ 0.8 µm: order code for "Measuring tube material", option SB, SE or Ra ≤ 0.4 µm: order code for "Measuring tube material", option SC, SF
Clamp
[in]
A
[mm]
B
[mm]
L
[mm]
Endress+Hauser 77
Page 78
Threaded couplings
L
A
B
Thread DIN 11851, DIN11864-1, SMS 1145
Length tolerance for dimension L in mm: +1.5 / –2.0
Thread DIN 11851, for pipe according to DIN11866, series A
1.4404 (316/316L) Order code for "Process connection", option FMW
DN
[mm]
8 Rd 34 × ¹⁄₈ 16 367
15 Rd 34 × ¹⁄₈ 16 398
25 Rd 52 × ¹⁄₆ 26 434
40 Rd 65 × ¹⁄₆ 38 560
50 Rd 78 × ¹⁄₆ 50 720
80 Rd 110 × ¹⁄₄ 81 900
100 Rd 130 × ¹⁄₄ 100 1 127
3-A version available: order code for "Additional approval", option LP in conjunction with Ra ≤ 0.8 µm: order code for "Measuring tube material", option SB, SE
A
[in]
B
[mm]
Proline Promass F 300
A0015628
L
[mm]
Thread DIN11864-1 Form A, for pipe according to DIN11866, series A
1.4404 (316/316L) Order code for "Process connection", option FLW
DN
[mm]
8 Rd 28 × ¹⁄₈ 10 367
15 Rd 34 × ¹⁄₈ 16 398
25 Rd 52 × ¹⁄₈ 26 434
40 Rd 65 × ¹⁄₆ 38 560
50 Rd 78 × ¹⁄₆ 50 720
80 Rd 110 × ¹⁄₄ 81 900
100 Rd 130 × ¹⁄₄ 100 1 127
3-A version available: order code for "Additional approval", option LP in conjunction with Ra ≤ 0.8 µm: order code for "Measuring tube material", option SB, SE or Ra ≤ 0.4 µm: order code for "Measuring tube material", option SC, SF
A
[in]
B
[mm]
L
[mm]
78 Endress+Hauser
Page 79
Proline Promass F 300
Thread SMS 1145
1.4404 (316/316L) Order code for "Process connection", option SCS
DN
[mm]
8 Rd 40 × ¹⁄₆ 22.6 367
15 Rd 40 × ¹⁄₆ 22.6 398
25 Rd 40 × ¹⁄₆ 22.6 434
40 Rd 60 × ¹⁄₆ 35.6 560
50 Rd 70 × ¹⁄₆ 48.6 720
80 Rd 98 × ¹⁄₆ 72.9 900
100 Rd 132 × ¹⁄₆ 97.6 1 127
3-A version available: order code for "Additional approval", option LP in conjunction with Ra ≤ 0.8 µm: order code for "Measuring tube material", option SB, SE
A
[in]
B
[mm]
L
[mm]
Endress+Hauser 79
Page 80
Thread ISO 2853
L
A
B
Length tolerance for dimension L in mm: +1.5 / –2.0
Thread ISO 2853, for pipe according to ISO 2037
1.4404 (316/316L) Order code for "Process connection", option JSF
DN
[mm]
8 37.13 22.6 367
15 37.13 22.6 398
25 37.13 22.6 434
40 52.68 35.6 560
50 64.16 48.6 720
80 91.19 72.9 900
100 118.21 97.6 1 127
3-A version available: order code for "Additional approval", option LP in conjunction with Ra ≤ 0.8 µm: order code for "Measuring tube material", option SB, SE or Ra ≤ 0.4 µm: order code for "Measuring tube material", option SC, SF
1)
A
[mm]
B
[mm]
Proline Promass F 300
A0015623
L
[mm]
1) Max. thread diameter as per ISO 2853 annex A
80 Endress+Hauser
Page 81
Proline Promass F 300
L
A
B
VCO
Length tolerance for dimension L in mm: +1.5 / –2.0
8-VCO-4 (½")
1.4404 (316/316L) Order code for "Process connection", option CVS
DN
[mm]
8 AF 1 10.2 390
A
[in]
B
[mm]
A0015624
L
[mm]
12-VCO-4 (¾")
1.4404 (316/316L) Order code for "Process connection", option CWS
DN
[mm]
15 AF 1½ 15.7 430
A
[in]
B
[mm]
L
[mm]
Endress+Hauser 81
Page 82
Accessories
DN 8 ( ")...150 (6")!⁄# DN 250 (10")
L L
45°
A
A
SW 1"
2
½" NPT
1
½" NPT
35 (1.38) ca. 75 (approx. 2.95)
1
1;2
1
1;2
146 (5.75)
48 (1.9)
12 (0.47)
280 (11.0) 255 (10.0)
134 (5.3)
30 (1.18)
Rupture disk/purge connections
Proline Promass F 300
 38
1 Connection nipple for purge connections: order code for "Sensor options", option CH "Purge connection" 2 Connection nipple with rupture disk: order code for "Sensor option", option CA "Rupture disk"
DN A L
[mm] [mm] [mm]
8 62 216
15 62 220
25 62 260
40 67 310
50 79 452
80 101 560
100 120 684
150 141 880
250 182 380
Protective cover
A0028914
82 Endress+Hauser
A0029553
Page 83
Proline Promass F 300
136 (5.35) 114 (4.49)
86 (3.39)
78 (3.07)
!103 (4.06)
!12 (0.47)
7.0 (0.27)
68 (2.7)105 (4.1)
173 (6.8)
Remote display and operating module DKX001
A0028921
 39 Engineering unit mm (in)
External WLAN antenna
The external WLAN antenna is not suitable for use in hygienic applications.
External WLAN antenna mounted on device
A0028923
 40 Engineering unit mm (in)
External WLAN antenna mounted with cable
The external WLAN antenna can be mounted separately from the transmitter if the transmission/ reception conditions at the transmitter mounting location are poor.
Endress+Hauser 83
Page 84
70 (2.8)
1500 (59.1)
72 (2.8)
 41 Engineering unit mm (in)
Proline Promass F 300
A0033597
84 Endress+Hauser
Page 85
Proline Promass F 300
D
L
E
F
A
CB
K
G
H
I
M

Dimensions in US units Compact version

Order code for "Housing", option A "Aluminum, coated"
DN A
[in] [in] [in] [in] [in] [in] [in] [in] [in] [in] [in] [in] [in]
³⁄₈ 6.65 2.68 3.98 2.95 10.22 13.17 7.87 2.32 5.55 0.211
½ 6.65 2.68 3.98 2.95 10.22 13.17 7.87 2.32 5.55 0.33
1 6.65 2.68 3.98 2.95 10.22 13.17 7.87 2.32 5.55 0.47
6.65 2.68 3.98 4.13 10.41 14.55 7.87 2.32 5.55 0.69
2 6.65 2.68 3.98 5.55 10.81 16.36 7.87 2.32 5.55 1.02
3 6.65 2.68 3.98 7.87 11.59 19.47 7.87 2.32 5.55 1.59
4 6.65 2.68 3.98 10.00 12.30 22.30 7.87 2.32 5.55 2.02
6 6.65 2.68 3.98 14.88 13.13 28.01 7.87 2.32 5.55 2.71
10 6.65 2.68 3.98 21.57 14.86 36.44 7.87 2.32 5.55 4.03
1)
A0033786
1)
B
C D E
2) 3)F 2) 3)G 4)
H I
4)
K L M
5)
2.76
5)
2.76
5)
2.76
5)
3.11
5)
3.90
5)
5.47
5)
6.93
5)
8.58
5)
12.01
1) Depending on the cable gland used: values up to + 1.18 in
2) With order code for "Sensor option", option CG or order code for "Measuring tube material", option SD, SE, SF, TH, LA: values +2.76 in
3) With order code for "Measuring tube material", option TT, TU: values +4.09 in
4) For version without local display: values - 1.18 in
5) Depends on the process connection in question
Order code for "Housing", option A "Aluminum, coated"; Ex d
DN A
1)
B
C D E
2) 3)
F G
4)
H I K L M
1)
[in] [in] [in] [in] [in] [in] [in] [in] [in] [in] [in] [in] [in]
³⁄₈ 7.40 3.35 4.06 2.95 11.42 14.37 8.11 2.28 5.83 0.211
½ 7.40 3.35 4.06 2.95 11.42 14.37 8.11 2.28 5.83 0.33
1 7.40 3.35 4.06 2.95 11.42 14.37 8.11 2.28 5.83 0.47
7.40 3.35 4.06 4.13 11.59 15.73 8.11 2.28 5.83 0.69
2 7.40 3.35 4.06 5.55 11.99 17.54 8.11 2.28 5.83 1.02
3 7.40 3.35 4.06 7.87 12.78 20.65 8.11 2.28 5.83 1.59
4 7.40 3.35 4.06 10 13.48 23.48 8.11 2.28 5.83 2.02
5)
2.76
5)
2.76
5)
2.76
5)
3.11
5)
3.90
5)
5.47
5)
6.93
Endress+Hauser 85
Page 86
Proline Promass F 300
1)
DN A
1)
B
C D E
[in] [in] [in] [in] [in] [in] [in] [in] [in] [in] [in] [in] [in]
6 7.40 3.35 4.06 14.88 14.31 29.19 8.11 2.28 5.83 2.71
10 7.40 3.35 4.06 21.57 16.04 37.62 8.11 2.28 5.83 4.03
1) Depending on the cable gland used: values up to + 1.18 in
2) With order code for "Sensor option", option CG or order code for "Measuring tube material", option SD, SE, SF, TH, LA: values +2.76 in
3) With order code for "Measuring tube material", option TT, TU: values +4.09 in
4) For version without local display: values - 1.49 in
5) Depends on the process connection in question
Order code for "Housing", option B "Stainless, hygienic"
1)
DN A
[in] [in] [in] [in] [in] [in] [in] [in] [in] [in] [in] [in] [in]
³⁄₈ 7.20 2.87 4.33 2.95 10.22 13.17 8.15 2.56 5.59 0.211
½ 7.20 2.87 4.33 2.95 10.22 13.17 8.15 2.56 5.59 0.33
1 7.20 2.87 4.33 2.95 10.22 13.17 8.15 2.56 5.59 0.47
7.20 2.87 4.33 4.13 10.41 14.55 8.15 2.56 5.59 0.69
2 7.20 2.87 4.33 5.55 10.81 16.36 8.15 2.56 5.59 1.02
3 7.20 2.87 4.33 7.87 11.59 19.47 8.15 2.56 5.59 1.59
4 7.20 2.87 4.33 10.00 12.30 22.30 8.15 2.56 5.59 2.02
6 7.20 2.87 4.33 14.88 13.13 28.01 8.15 2.56 5.59 2.71
10 7.20 2.87 4.33 21.57 14.86 36.44 8.15 2.56 5.59 4.03
1)
B
C D E
2) 3)
2) 3)
F G
F G
4)
H I K L M
4)
H I K L M
5)
8.58
5)
12.01
5)
2.76
5)
2.76
5)
2.76
5)
3.11
5)
3.90
5)
5.47
5)
6.93
5)
8.58
5)
12.01
1) Depending on the cable gland used: values up to + 1.18 in
2) With order code for "Sensor option", option CG or order code for "Measuring tube material", option SD, SE, SF, TH, LA: values +2.76 in
3) With order code for "Measuring tube material", option TT, TU: values +4.09 in
4) For version without local display: values - 0.51 in
5) Depends on the process connection in question
Order code for "Housing", option L "Cast, stainless"
1)
DN A
1)
B
C D E
[in] [in] [in] [in] [in] [in] [in] [in] [in] [in] [in] [in] [in]
³⁄₈ 7.32 3.35 3.98 2.95 11.42 14.37 8.7 2.48 6.22 0.211
½ 7.32 3.35 3.98 2.95 11.42 14.37 8.7 2.48 6.22 0.33
1 7.32 3.35 3.98 2.95 11.42 14.37 8.7 2.48 6.22 0.47
7.32 3.35 3.98 4.13 11.59 15.73 8.7 2.48 6.22 0.69
2 7.32 3.35 3.98 5.55 11.99 17.54 8.7 2.48 6.22 1.02
3 7.32 3.35 3.98 7.87 12.78 20.65 8.7 2.48 6.22 1.59
4 7.32 3.35 3.98 10 13.48 23.48 8.7 2.48 6.22 2.02
6 7.32 3.35 3.98 14.88 14.31 29.19 8.7 2.48 6.22 2.71
10 7.32 3.35 3.98 21.57 16.04 37.62 8.7 2.48 6.22 4.03
1) Depending on the cable gland used: values up to + 1.18 in
2) With order code for "Sensor option", option CG or order code for "Measuring tube material", option SD, SE, SF, TH, LA: values +2.76 in
3) With order code for "Measuring tube material", option TT, TU: values +4.09 in
4) Depends on the process connection in question
2) 3)
F G H I K L M
4)
2.76
4)
2.76
4)
2.76
4)
3.11
4)
3.90
4)
5.47
4)
6.93
4)
8.58
4)
12.01
86 Endress+Hauser
Page 87
Proline Promass F 300
C
D
A
B
E
L
Flange connections
Fixed flange ASME B16.5
A0015621
Length tolerance for dimension L in inch:
• DN ≤ 4": +0.06 / –0.08
• DN ≥ 5": +0.14
Flange according to ASME B16.5: Class 150
1.4404 (F316/F316L): order code for "Process connection", option AAS Alloy C22: order code for "Process connection", option AAC
DN
[in]
1)
³⁄₈
½ 3.54 2.37 4 × ⌀0.62 0.44 0.62 15.91
1 4.33 3.13 4 × ⌀0.62 0.56 1.05 17.32
4.92 3.87 4 × ⌀0.62 0.69 1.61 21.65
2 5.91 4.75 4 × ⌀0.75 0.75 2.07 28.15
3 7.48 6.00 4 × ⌀0.75 0.94 3.07 33.07
4 9.06 7.50 8 × ⌀0.75 0.94 4.03 44.37
6 11.02 9.50 8 × ⌀0.88 1 6.07 55.04
10 15.94 14.25 12 × ⌀1.0 1.19 10.02 72.13
Surface roughness (flange): Ra 125 to 250 µin
1) DN ³⁄₈" with DN ½" flanges as standard
Flange according to ASME B16.5: Class 150 with reduction in nominal diameter
1.4404 (F316/F316L)
DN
[in]
2 AHS 5.91 4.75 4 × ⌀ 0.75 0.75 2.07 21.65
3 2 AJS 7.48 6 4 × ⌀ 0.75 0.94 3.07 28.35
4 3 ALS 9.06 7.5 8 × ⌀ 0.75 0.94 4.03 34.41
6 4 ANS 11.02 9.5 8 × ⌀ 0.88 1 6.07 45.94
8 6 APS 13.58 11.75 8 × ⌀ 0.88 1.14 7.98 57.52
Surface roughness (flange): Ra 125 to 250 µin
A
[in]
3.54 2.37 4 × ⌀ 0.62 0.44 0.62 14.57
reduction
to DN
[in]
Order code for
B
[in]
"Process
connection",
option
A
[in]
C
[in]
B
[in]
D
[in]
C
[in]
E
[in]
D
[in]
E
[in]
L
[in]
L
[in]
Endress+Hauser 87
Page 88
Proline Promass F 300
Flange according to ASME B16.5: Class 300
1.4404 (F316/F316L): order code for "Process connection", option ABS Alloy C22: order code for "Process connection", option ABC
DN
[in]
1)
³⁄₈
½ 3.74 2.63 4 × ⌀0.62 0.56 0.62 15.91
1 4.92 3.50 4 × ⌀0.75 0.69 1.05 17.32
6.10 4.50 4 × ⌀0.88 0.81 1.61 21.65
2 6.50 5.00 8 × ⌀0.75 0.88 2.07 28.15
3 8.27 6.63 8 × ⌀0.88 1.12 3.07 33.07
4 10.04 7.87 8 × ⌀0.88 1.25 4.03 44.37
6 12.6 10.63 12 × ⌀0.88 1.44 6.07 55.79
10 17.52 15.25 16 × ⌀1.12 1.87 10.02 73.35
Surface roughness (flange): Ra 125 to 250 µin
1) DN ³⁄₈" with DN ½" flanges as standard
A
[in]
B
[in]
C
[in]
D
[in]
E
[in]
3.74 2.63 4 × ⌀0.62 0.56 0.62 14.57
L
[in]
Flange according to ASME B16.5: Class 300 with reduction in nominal diameter
1.4404 (F316/F316L)
DN
[in]
reduction
to DN
[in]
Order code for
"Process
connection",
A
[in]
B
[in]
C
[in]
D
[in]
E
[in]
option
2 AIS 6.5 5 8 × ⌀ 0.75 0.88 2.07 24.21
3 2 AKS 8.27 6.63 8 × ⌀ 0.88 1.12 3.07 28.82
4 3 AMS 10.04 7.87 8 × ⌀ 0.88 1.25 4.03 35.2
6 4 AOS 12.6 10.63 12 × ⌀ 0.88 1.44 6.07 46.73
8 6 AQS 14.96 13 12 × ⌀ 1 1.64 7.98 57.52
Surface roughness (flange): Ra 125 to 250 µin
Flange according to ASME B16.5: Class 600
1.4404 (F316/F316L): order code for "Process connection", option ACS Alloy C22: order code for "Process connection", option ACC
DN
[in]
³⁄₈
1)
A
[in]
B
[in]
C
[in]
D
[in]
E
[in]
3.74 2.63 4 × ⌀0.62 0.81 0.55 15.75
½ 3.74 2.63 4 × ⌀0.62 0.81 0.55 16.54
1 4.92 3.50 4 × ⌀0.75 0.94 0.96 19.29
6.10 4.50 4 × ⌀0.88 1.13 1.5 23.62
2 6.50 5.00 8 × ⌀0.75 1.25 1.94 29.21
3 8.27 6.63 8 × ⌀0.88 1.5 2.9 35.43
4 10.83 8.50 8 × ⌀1.00 1.91 3.83 45.55
6 13.98 11.50 12 × ⌀1.12 1.88 6.07 57.76
10 20.08 17.00 16 × ⌀1.38 2.75 10.02 76.61
Surface roughness (flange): Ra 125 to 250 µin
L
[in]
L
[in]
1) DN ³⁄₈" with DN ½" flanges as standard
88 Endress+Hauser
Page 89
Proline Promass F 300
L
C
D
A
B
F
E
Lap joint flange ASME B16.5
Length tolerance for dimension L in inch: +0.06 / –0.08
Lap joint flange according to ASME B16.5: Class 150
1.4301 (F304), wetted parts Alloy C22 Order code for "Process connection", option ADC
DN
[in]
2)
³⁄₈
½ 3.54 2.37 4 × ⌀ 0.62 0.59 1.38 0.62 15.91 0
1 4.33 3.13 4 × ⌀ 0.62 0.63 2 1.05 17.32 0
4.92 3.87 4 × ⌀ 0.62 0.63 2.88 1.61 21.65 0
2 5.91 4.75 4 × ⌀ 0.75 0.75 3.62 2.07 28.15 0
3 7.48 6.00 4 × ⌀ 0.75 0.88 5 3.07 33.07 0
4 9.06 7.50 8 × ⌀ 0.75 1.02 6.19 4.03 44.37 0
Surface roughness (flange): Ra 125 to 492 µin
A
[in]
B
[in]
C
[in]
D
[in]
E
[in]
F
[in]
L
[in]
3.54 2.37 4 × ⌀ 0.62 0.59 1.38 0.62 14.57 0
L
diff
[in]
A0022221
1)
1) Difference to installation length of the welding neck flange (order code for "Process connection", option AAC)
2) DN ³⁄₈" with DN ½" flanges as standard
Lap joint flange according to ASME B16.5: Class 300
1.4301 (F304), wetted parts Alloy C22 Order code for "Process connection", option AEC
DN
[in]
³⁄₈
A
[in]
2)
3.74 2.63 4 × ⌀ 0.62 0.65 1.38 0.62 14.8 +0.23
B
[in]
C
[in]
D
[in]
E
[in]
F
[in]
L
[in]
L
diff
[in]
1)
½ 3.74 2.63 4 × ⌀ 0.62 0.65 1.38 0.62 15.98 +0.07
1 4.92 3.50 4 × ⌀ 0.75 0.83 2 1.05 17.72 +0.40
6.10 4.50 4 × ⌀ 0.88 0.91 2.88 1.61 22.2 +0.55
2 6.50 5.00 8 × ⌀ 0.75 1 3.62 2.07 28.23 +0.08
3 8.27 6.63 8 × ⌀ 0.88 1.22 5 3.07 33.57 +0.50
Endress+Hauser 89
Page 90
Proline Promass F 300
Lap joint flange according to ASME B16.5: Class 300
1.4301 (F304), wetted parts Alloy C22 Order code for "Process connection", option AEC
DN
[in]
4 10.04 7.87 8 × ⌀ 0.88 1.26 6.19 4.03 44.84 +0.47
Surface roughness (flange): Ra 125 to 492 µin
1) Difference to installation length of the welding neck flange (order code for "Process connection", option AAC)
2) DN ³⁄₈" with DN ½" flanges as standard
Lap joint flange according to ASME B16.5, Class 600
1.4301 (F304), wetted parts Alloy C22 Order code for "Process connection", option AFC
DN
[in]
2)
³⁄₈
½ 3.74 2.63 4 × ⌀ 0.62 0.67 1.38 0.55 16.54 0
1 4.92 3.50 4 × ⌀ 0.75 0.85 2 0.96 19.29 0
6.10 4.50 4 × ⌀ 0.88 0.98 2.88 1.5 23.62 0
2 6.50 5.00 8 × ⌀ 0.75 1.1 3.62 1.94 29.21 0
3 8.27 6.63 8 × ⌀ 0.88 1.38 5 2.9 35.43 0
4 10.83 8.50 8 × ⌀ 1 1.73 6.19 3.83 45.94 +0.39
Surface roughness (flange): Ra 125 to 492 µin
A
[in]
A
[in]
B
[in]
B
[in]
C
[in]
C
[in]
D
[in]
D
[in]
E
[in]
E
[in]
F
[in]
F
[in]
L
[in]
L
[in]
L
[in]
L
[in]
3.74 2.63 4 × ⌀ 0.62 0.67 1.38 0.55 15.75 0
diff
diff
1)
1)
1) Difference to installation length of the welding neck flange (order code for "Process connection", option AAC)
2) DN ³⁄₈" with DN ½" flanges as standard
90 Endress+Hauser
Page 91
Proline Promass F 300
L
A
B
Clamp connections
Tri-Clamp
Length tolerance for dimension L in inch: +0.06 / –0.08
Tri-Clamp (½"), DIN 11866 series C
1.4404 (316/316L) Order code for "Process connection", option FDW
DN
[in]
³⁄₈ ½ 0.98 0.37 14.4
½ ½ 0.98 0.37 15.7
3-A version available: order code for "Additional approval", option LP in conjunction with Ra ≤ 32 µin: order code for "Measuring tube material", option SB, SE or Ra ≤ 16 µin: order code for "Measuring tube material", option SC, SF
Clamp
[in]
A
[in]
B
[in]
A0015625
L
[in]
Tri-Clamp ( ≥ 1"), DIN 11866 series C
1.4404 (316/316L) Order code for "Process connection", option FTS
DN
[in]
³⁄₈ 1 1.98 0.87 14.4
½ 1 1.98 0.87 15.7
1 1 1.98 0.87 17.1
1.98 1.37 22.0
2 2 2.52 1.87 28.3
3 3 3.58 2.87 35.4
4 4 4.68 3.83 44.4
3-A version available: order code for "Additional approval", option LP in conjunction with Ra ≤ 32 µin: order code for "Measuring tube material", option SB, SE or Ra ≤ 16 µin: order code for "Measuring tube material", option SC, SF
Clamp
[in]
A
[in]
B
[in]
L
[in]
Endress+Hauser 91
Page 92
Threaded couplings
L
A
B
Thread SMS 1145
Length tolerance for dimension L in inch: +0.06 / –0.08
Thread SMS 1145
1.4404 (316/316L) Order code for "Process connection", option SCS
DN
[in]
³⁄₈ Rd 40 × ¹⁄₆ 0.89 14.45
½ Rd 40 × ¹⁄₆ 0.89 15.67
1 Rd 40 × ¹⁄₆ 0.89 17.09
Rd 60 × ¹⁄₆ 1.4 22.05
2 Rd 70 × ¹⁄₆ 1.91 28.35
3 Rd 98 × ¹⁄₆ 2.87 35.43
4 Rd 132 × ¹⁄₆ 3.84 44.37
3-A version available: order code for "Additional approval", option LP in conjunction with Ra ≤ 32 µin: order code for "Measuring tube material", option SB, SE
A
[in]
B
[in]
Proline Promass F 300
A0015628
L
[in]
92 Endress+Hauser
Page 93
Proline Promass F 300
L
A
B
DN 8 ( ")...150 (6")!⁄# DN 250 (10")
L L
45°
A
A
SW 1"
2
½" NPT
1
½" NPT
35 (1.38) ca. 75 (approx. 2.95)
1
1;2
1
1;2
VCO
Length tolerance for dimension L in inch: +0.06 / –0.08
8-VCO-4 (½")
1.4404 (316/316L) Order code for "Process connection", option CVS
DN
[in]
³⁄₈ AF 1 0.4 15.35
A
[in]
B
[in]
A0015624
L
[in]
12-VCO-4 (¾")
1.4404 (316/316L) Order code for "Process connection", option CWS
Accessories
Rupture disk/purge connections
DN
[in]
½ AF 1½ 0.62 16.93
A
[in]
B
[in]
L
[in]
Endress+Hauser 93
1 Connection nipple for purge connections: order code for "Sensor options", option CH "Purge connection" 2 Connection nipple with rupture disk: order code for "Sensor option", option CA "Rupture disk"
A0028914
Page 94
Protective cover
146 (5.75)
48 (1.9)
12 (0.47)
280 (11.0) 255 (10.0)
134 (5.3)
30 (1.18)
136 (5.35) 114 (4.49)
86 (3.39)
78 (3.07)
!103 (4.06)
!12 (0.47)
7.0 (0.27)
Proline Promass F 300
DN A L
[in] [in] [in]
³⁄₈ 2.44 8.50
½ 2.44 8.66
1 2.44 10.24
2.64 12.20
2 3.11 17.78
3 3.98 22.0
4 4.72 27.0
6 5.55 34.6
10 7.17 14.96
Remote display and operating module DKX001
 42 Engineering unit mm (in)
External WLAN antenna
The external WLAN antenna is not suitable for use in hygienic applications.
A0029553
A0028921
94 Endress+Hauser
Page 95
Proline Promass F 300
68 (2.7)105 (4.1)
173 (6.8)
70 (2.8)
1500 (59.1)
72 (2.8)
External WLAN antenna mounted on device
A0028923
 43 Engineering unit mm (in)
External WLAN antenna mounted with cable
The external WLAN antenna can be mounted separately from the transmitter if the transmission/ reception conditions at the transmitter mounting location are poor.
A0033597
 44 Engineering unit mm (in)

Weight

All values (weight exclusive of packaging material) refer to devices with EN/DIN PN 40 flanges. Weight specifications including transmitter as per order code for "Housing", option A "Aluminum, coated".
Different values due to different transmitter versions:
• Transmitter version for the hazardous area (Order code for "Housing", option A "Aluminum, coated"; Ex d): +2 kg (+4.4 lbs)
• Cast transmitter version, stainless (Order code for "Housing", option L "Cast, stainless"): +6 kg (+13 lbs)
• Transmitter version for hygienic area (Order code for "Housing", option B "Stainless, hygienic"): +0.2 kg (+0.44 lbs)
Weight in SI units
DN [mm] Weight [kg]
8 11
15 12
25 14
40 19
Endress+Hauser 95
50 30
Page 96
DN [mm] Weight [kg]
Weight in US units
DN [in] Weight [lbs]
Proline Promass F 300
80 55
100 96
150 154
250 400
3/8 24
½ 26
1 31
42
2 66
3 121
4 212
6 340
10 882

Materials Transmitter housing

Order code for "Housing":
• Option A "Aluminum, coated": aluminum, AlSi10Mg, coated
• Option B "Stainless, hygienic": stainless steel, 1.4404 (316L)
• Option L "Cast, stainless": cast, stainless steel, 1.4409 (CF3M) similar to 316L
Window material
Order code for "Housing":
• Option A "Aluminum, coated": glass
• Option B "Stainless, hygienic": polycarbonate
• Option L "Cast, stainless": glass
Seals
Order code for "Housing": Option B "Stainless, hygienic": EPDM and silicone
96 Endress+Hauser
Page 97
Proline Promass F 300
1
2
3
4
Cable entries/cable glands
A0028352
 45 Possible cable entries/cable glands
1 Female thread M20 × 1.5 2 Cable gland M20 × 1.5 3 Adapter for cable entry with internal thread G ½" or NPT ½" 4 Device plugs
Order code for "Housing", option A "Aluminum, coated"
The various cable entries are suitable for hazardous and non-hazardous areas.
Cable entry/cable gland Material
Cable gland M20 × 1.5 Plastic/nickel-plated brass
Adapter for cable entry with internal thread G ½" Nickel-plated brass
Adapter for cable entry with internal thread NPT ½"
Device plug Plug M12 × 1
• Socket: Stainless steel, 1.4404 (316L)
• Contact housing: Polyamide
• Contacts: Gold-plated brass
Order code for "Housing", option B "Stainless, hygienic"
The various cable entries are suitable for hazardous and non-hazardous areas.
Cable entry/cable gland Material
Cable gland M20 × 1.5 Plastic
Adapter for cable entry with internal thread G ½" Nickel-plated brass
Adapter for cable entry with internal thread NPT ½"
Device plug Plug M12 × 1
• Socket: Stainless steel, 1.4404 (316L)
• Contact housing: Polyamide
• Contacts: Gold-plated brass
Order code for "Housing", option L "Cast, stainless"
The various cable entries are suitable for hazardous and non-hazardous areas.
Cable entry/cable gland Material
Cable gland M20 × 1.5 Stainless steel, 1.4404 (316L)
Adapter for cable entry with internal thread G ½"
Endress+Hauser 97
Page 98
Proline Promass F 300
Cable entry/cable gland Material
Adapter for cable entry with internal thread NPT ½"
Device plug Plug M12 × 1
• Socket: Stainless steel, 1.4404 (316L)
• Contact housing: Polyamide
• Contacts: Gold-plated brass
Device plug
Electrical connection Material
Plug M12x1 • Socket: Stainless steel, 1.4404 (316L)
• Contact housing: Polyamide
• Contacts: Gold-plated brass
Sensor housing
The material of the sensor housing depends on the option selected in the order code for "Measuring tube mat., wetted surface".
Order code for "Measuring tube mat., wetted surface"
Option HA, SA, SD, TH • Acid and alkali-resistant outer surface
Option SB, SC, SE, SF • Acid and alkali-resistant outer surface
Option TT, TU, LA • Acid and alkali-resistant outer surface
Material
• Stainless steel 1.4301 (304)
With order code for "Sensor option", option CC
"316L Sensor housing": stainless steel, 1.4404 (316L)
• Stainless steel 1.4301 (304)
• Stainless steel, 1.4404 (316L)
Measuring tubes
• DN 8 to 100 (3/8…4"): stainless steel, 1.4539 (904L); Manifold: stainless steel, 1.4404 (316/316L)
• DN 150 (6"), DN 250 (10"): stainless steel, 1.4404 (316/316L); Manifold: stainless steel, 1.4404 (316/316L)
• DN 8 to 250 (3/8 to 10"): Alloy C22, 2.4602 (UNS N06022); Manifold: Alloy C22, 2.4602 (UNS N06022)
High-temperature version
DN 25, DN 50, DN 80 (DN 1", DN 2", DN 3"): Alloy C22, 2.4602 (UNS N06022)
Process connections
• Flanges according to EN 1092-1 (DIN2501) / according to ASME B 16.5 / as per JIS B2220: – Stainless steel, 1.4404 (F316/F316L) – Alloy C22, 2.4602 (UNS N06022) – Lap joint flanges: stainless steel, 1.4301 (F304); wetted parts Alloy C22
• All other process connections: Stainless steel, 1.4404 (316/316L)
High-temperature version
Flanges according to EN 1092-1 (DIN2501) / according to ASME B 16.5 / as per JIS B2220: – Stainless steel, 1.4404 (F316/F316L) – Alloy C22, 2.4602 (UNS N06022)
Available process connections→  99
Seals
Welded process connections without internal seals
98 Endress+Hauser
Page 99
Proline Promass F 300
Accessories
Protective cover
Stainless steel, 1.4404 (316L)
External WLAN antenna
• Antenna: ASA plastic (acrylic ester-styrene-acrylonitrile) and nickel-plated brass
• Adapter: Stainless steel and nickel-plated brass
• Cable: Polyethylene
• Plug: Nickel-plated brass
• Angle bracket: Stainless steel

Process connections

Surface roughness

• Fixed flange connections: – EN 1092-1 (DIN 2501) flange – EN 1092-1 (DIN 2512N) flange – Namur lengths in accordance with NE 132 – ASME B16.5 flange – JIS B2220 flange – DIN 11864-2 Form A flange, DIN 11866 series A, flange with notch
• Clamp connections: Tri-Clamp (OD tubes), DIN 11866 series C
• Thread: – DIN 11851 thread, DIN 11866 series A – SMS 1145 thread – ISO 2853 thread, ISO 2037 – DIN 11864-1 Form A thread, DIN 11866 series A
• VCO connections: – 8-VCO-4 – 12-VCO-4
Process connection materials →  98
All data relate to parts in contact with fluid. The following surface roughness quality can be ordered.
• Not polished
• Ra
= 0.8 µm (32 µin)
max
• Ra
= 0.4 µm (16 µin)
max

Operability

Operating concept

Endress+Hauser 99
Operator-oriented menu structure for user-specific tasks
• Commissioning
• Operation
• Diagnostics
• Expert level
Fast and safe commissioning
• Guided menus ("Make-it-run" wizards) for applications
• Menu guidance with brief descriptions of the individual parameter functions
• Device access via Web server or SmartBlue app →  119
• WLAN access to the device via mobile handheld terminal, tablet or smart phone
Reliable operation
• Operation in local language →  100
• Uniform operating philosophy applied to device and operating tools
• If replacing electronic modules, transfer the device configuration via the integrated memory (HistoROM backup) which contains the process and measuring device data and the event logbook. No need to reconfigure.
Efficient diagnostics increase measurement availability
• Troubleshooting measures can be called up via the device and in the operating tools
• Diverse simulation options, logbook for events that occur and optional line recorder functions
Page 100
Proline Promass F 300

Languages

Can be operated in the following languages:
• Via local operation English, German, French, Spanish, Italian, Dutch, Portuguese, Polish, Russian, Turkish, Chinese, Japanese, Korean, Bahasa (Indonesian), Vietnamese, Czech, Swedish
• Via Web browser English, German, French, Spanish, Italian, Dutch, Portuguese, Polish, Russian, Turkish, Chinese, Japanese, Korean, Bahasa (Indonesian), Vietnamese, Czech, Swedish
• Via "FieldCare", "DeviceCare" operating tool: English, German, French, Spanish, Italian, Chinese, Japanese

Local operation Via display module

Two display modules are available:
• Order code for "Display; operation", option F "4-line, illuminated, graphic display; touch control"
• Order code for "Display; operation", option G "4-line, illuminated, graphic display; touch control + WLAN"
Information about WLAN interface →  107
 46 Operation with touch control
Display elements
• 4-line, illuminated, graphic display
• White background lighting; switches to red in event of device errors
• Format for displaying measured variables and status variables can be individually configured
• Permitted ambient temperature for the display: –20 to +60 °C (–4 to +140 °F) The readability of the display may be impaired at temperatures outside the temperature range.
Operating elements
• External operation via touch control (3 optical keys) without opening the housing: , , 
• Operating elements also accessible in the various zones of the hazardous area
Via remote display and operating module DKX001
The remote display and operating module DKX001 is available as an optional extra →  117.
• The remote display and operating module DKX001 is only available for the following housing versions, order code for "Housing": – Option A "Aluminum, coated" – Option L "Cast, stainless"
• The measuring device is always supplied with a dummy cover when the remote display and operating module DKX001 is ordered directly with the measuring device. Display or operation at the transmitter is not possible in this case.
• If ordered subsequently, the remote display and operating module DKX001 may not be connected at the same time as the existing measuring device display module. Only one display or operation unit may be connected to the transmitter at any one time.
A0026785
100 Endress+Hauser
Loading...