Endress+Hauser F 200 Operating Manual

Page 1
BA01112D/06/EN/05.16 71340679
Valid as of version
Products Solutions Services
Operating Instructions
Proline Promass F 200 HART
Coriolis flowmeter
Page 2
• Make sure the document is stored in a safe place such that it is always available when working on or with the device.
• To avoid danger to individuals or the facility, read the "Basic safety instructions" section carefully, as well as all other safety instructions in the document that are specific to working procedures.
• The manufacturer reserves the right to modify technical data without prior notice. Your Endress+Hauser Sales Center will supply you with current information and updates to these instructions.
2 Endress+Hauser
Page 3

Table of contents

Table of contents
1 Document information .............. 6
1.1 Document function ..................... 6
1.2 Symbols used .......................... 6
1.2.1 Safety symbols .................. 6
1.2.2 Electrical symbols ................ 6
1.2.3 Tool symbols .................... 7
1.2.4 Symbols for certain types of
information .................... 7
1.2.5 Symbols in graphics ............... 7
1.3 Documentation ........................ 8
1.3.1 Standard documentation ........... 8
1.3.2 Supplementary device-dependent
documentation .................. 8
1.4 Registered trademarks ................... 8
2 Basic safety instructions ............ 9
2.1 Requirements for the personnel ............ 9
2.2 Designated use ........................ 9
2.3 Workplace safety ...................... 10
2.4 Operational safety ..................... 10
2.5 Product safety ........................ 10
2.6 IT security ........................... 11
2.7 Device-specific IT security ................ 11
2.7.1 Protecting access via hardware write
protection ..................... 11
2.7.2 Protecting access via a password .... 11
2.7.3 Access via fieldbus ............... 11
3 Product description ................ 12
3.1 Product design ........................ 12
4 Incoming acceptance and product
identification ..................... 13
4.1 Incoming acceptance ................... 13
4.2 Product identification ................... 13
4.2.1 Transmitter nameplate ........... 14
4.2.2 Sensor nameplate ............... 15
4.2.3 Symbols on measuring device ...... 16
5 Storage and transport ............. 17
5.1 Storage conditions ..................... 17
5.2 Transporting the product ................ 17
5.2.1 Measuring devices without lifting
lugs ......................... 17
5.2.2 Measuring devices with lifting lugs .. 18
5.2.3 Transporting with a fork lift ........ 18
5.3 Packaging disposal ..................... 18
6 Installation ....................... 19
6.1 Installation conditions .................. 19
6.1.1 Mounting position ............... 19
6.1.2 Requirements from environment and
process ....................... 21
6.1.3 Special mounting instructions ...... 24
6.2 Mounting the measuring device ........... 24
6.2.1 Required tools .................. 24
6.2.2 Preparing the measuring device ..... 25
6.2.3 Mounting the measuring device ..... 25
6.2.4 Turning the transmitter housing .... 25
6.2.5 Turning the display module ........ 26
6.3 Post-installation check .................. 26
7 Electrical connection .............. 27
7.1 Connection conditions .................. 27
7.1.1 Required tools .................. 27
7.1.2 Connecting cable requirements ..... 27
7.1.3 Terminal assignment ............. 28
7.1.4 Requirements for the supply unit .... 28
7.1.5 Preparing the measuring device ..... 29
7.2 Connecting the measuring device .......... 30
7.2.1 Connecting the transmitter ........ 30
7.2.2 Ensuring potential equalization ..... 31
7.3 Special connection instructions ............ 31
7.3.1 Connection examples ............. 31
7.4 Ensuring the degree of protection .......... 33
7.5 Post-connection check .................. 34
8 Operation options ................. 35
8.1 Overview of operation options ............ 35
8.2 Structure and function of the operating
menu .............................. 36
8.2.1 Structure of the operating menu .... 36
8.2.2 Operating philosophy ............ 37
8.3 Access to the operating menu via the local
display ............................. 38
8.3.1 Operational display .............. 38
8.3.2 Navigation view ................ 39
8.3.3 Editing view ................... 41
8.3.4 Operating elements .............. 43
8.3.5 Opening the context menu ......... 43
8.3.6 Navigating and selecting from list ... 45
8.3.7 Calling the parameter directly ...... 45
8.3.8 Calling up help text .............. 46
8.3.9 Changing the parameters ......... 47
8.3.10 User roles and related access
authorization .................. 48
8.3.11 Disabling write protection via access
code ......................... 48
8.3.12 Enabling and disabling the keypad
lock ......................... 48
8.4 Access to the operating menu via the
operating tool ........................ 49
8.4.1 Connecting the operating tool ...... 50
8.4.2 Field Xpert SFX350, SFX370 ....... 51
8.4.3 FieldCare ..................... 51
Endress+Hauser 3
Page 4
Table of contents
8.4.4 DeviceCare .................... 52
8.4.5 AMS Device Manager ............ 52
8.4.6 SIMATIC PDM .................. 53
8.4.7 Field Communicator 475 .......... 53
9 System integration ................ 54
9.1 Overview of device description files ......... 54
9.1.1 Current version data for the device ... 54
9.1.2 Operating tools ................. 54
9.2 Measured variables via HART protocol ...... 54
9.3 Other settings ........................ 56
10 Commissioning .................... 59
10.1 Function check ....................... 59
10.2 Switching on the measuring device ......... 59
10.3 Setting the operating language ............ 59
10.4 Configuring the measuring device .......... 59
10.4.1 Defining the tag name ............ 60
10.4.2 Selecting and setting the medium ... 62
10.4.3 Setting the system units .......... 62
10.4.4 Configuring the current output ..... 65
10.4.5 Configuring the pulse/frequency/
switch output .................. 66
10.4.6 Configuring the local display ....... 73
10.4.7 Configuring the output
conditioning ................... 75
10.4.8 Configuring the low flow cut off ..... 78
10.4.9 Configuring the partial filled pipe
detection ..................... 79
10.5 Advanced settings ..................... 80
10.5.1 Carrying out a sensor adjustment .... 81
10.5.2 Configuring the totalizer .......... 82
10.5.3 Carrying out additional display
configurations .................. 84
10.5.4 Using parameters for device
administration ................. 86
10.6 Configuration management .............. 87
10.6.1 Function scope of the "Configuration
management" parameter .......... 88
10.7 Simulation ........................... 88
10.8 Protecting settings from unauthorized
access .............................. 90
10.8.1 Write protection via access code ..... 91
10.8.2 Write protection via write protection
switch ........................ 91
11 Operation ......................... 94
11.1 Reading the device locking status .......... 94
11.2 Adjusting the operating language .......... 94
11.3 Configuring the display ................. 94
11.4 Reading measured values ................ 94
11.4.1 Process variables ................ 94
11.4.2 "Totalizer" submenu .............. 95
11.4.3 Output values .................. 96
11.5 Adapting the measuring device to the process
conditions ........................... 97
11.6 Performing a totalizer reset .............. 97
11.6.1 Function scope of the "Control
Totalizer" parameter ............. 98
11.6.2 Function scope of the "Reset all
totalizers" parameter ............. 98
11.7 Showing data logging ................... 98
12 Diagnostics and troubleshooting .. 101
12.1 General troubleshooting ................ 101
12.2 Diagnostic information on local display ..... 103
12.2.1 Diagnostic message ............. 103
12.2.2 Calling up remedial measures ..... 105
12.3 Diagnostic information in DeviceCare or
FieldCare ........................... 105
12.3.1 Diagnostic options .............. 105
12.3.2 Calling up remedy information .... 107
12.4 Adapting the diagnostic information ...... 107
12.4.1 Adapting the diagnostic behavior ... 107
12.4.2 Adapting the status signal ........ 108
12.5 Overview of diagnostic information ....... 109
12.6 Pending diagnostic events .............. 112
12.7 Diagnostic list ....................... 112
12.8 Event logbook ....................... 113
12.8.1 Event history .................. 113
12.8.2 Filtering the event logbook ....... 113
12.8.3 Overview of information events .... 114
12.9 Resetting the measuring device .......... 115
12.9.1 Function scope of the "Device reset"
parameter .................... 115
12.10 Device information ................... 115
12.11 Firmware history ..................... 117
13 Maintenance .................... 119
13.1 Maintenance tasks .................... 119
13.1.1 Exterior cleaning ............... 119
13.1.2 Interior cleaning ............... 119
13.2 Measuring and test equipment ........... 119
13.3 Endress+Hauser services ............... 119
14 Repairs .......................... 120
14.1 General notes ....................... 120
14.1.1 Repair and conversion concept ..... 120
14.1.2 Notes for repair and conversion .... 120
14.2 Spare parts ......................... 120
14.3 Endress+Hauser services ............... 121
14.4 Return ............................. 121
14.5 Disposal ........................... 121
14.5.1 Removing the measuring device .... 121
14.5.2 Disposing of the measuring device .. 122
15 Accessories ...................... 123
15.1 Device-specific accessories .............. 123
15.1.1 For the transmitter ............. 123
15.1.2 For the sensor ................. 124
15.2 Communication-specific accessories ....... 124
15.3 Service-specific accessories .............. 125
4 Endress+Hauser
Page 5
15.4 System components ................... 125
16 Technical data ................... 127
16.1 Application ......................... 127
16.2 Function and system design ............. 127
16.3 Input .............................. 128
16.4 Output ............................ 129
16.5 Power supply ........................ 132
16.6 Performance characteristics ............. 133
16.7 Installation ......................... 137
16.8 Environment ........................ 137
16.9 Process ............................ 138
16.10 Mechanical construction ............... 140
16.11 Operability ......................... 143
16.12 Certificates and approvals .............. 145
16.13 Application packages .................. 147
16.14 Accessories ......................... 147
16.15 Documentation ...................... 148
Index ................................. 150
Table of contents
Endress+Hauser 5
Page 6
Document information Proline Promass F 200 HART
DANGER
WARNING
CAUTION
NOTICE

1 Document information

1.1 Document function

These Operating Instructions contain all the information that is required in various phases of the life cycle of the device: from product identification, incoming acceptance and storage, to mounting, connection, operation and commissioning through to troubleshooting, maintenance and disposal.

1.2 Symbols used

1.2.1 Safety symbols

Symbol Meaning
DANGER!
This symbol alerts you to a dangerous situation. Failure to avoid this situation will result in serious or fatal injury.
WARNING!
This symbol alerts you to a dangerous situation. Failure to avoid this situation can result in serious or fatal injury.
CAUTION!
This symbol alerts you to a dangerous situation. Failure to avoid this situation can result in minor or medium injury.
NOTE!
This symbol contains information on procedures and other facts which do not result in personal injury.

1.2.2 Electrical symbols

Symbol Meaning
Direct current
Alternating current
Direct current and alternating current
Ground connection
A grounded terminal which, as far as the operator is concerned, is grounded via a grounding system.
Protective ground connection
A terminal which must be connected to ground prior to establishing any other connections.
Equipotential connection
A connection that has to be connected to the plant grounding system: This may be a potential equalization line or a star grounding system depending on national or company codes of practice.
6 Endress+Hauser
Page 7
Proline Promass F 200 HART Document information
A
1.
1.
-
.

1.2.3 Tool symbols

Symbol Meaning
Flat blade screwdriver
Allen key
Open-ended wrench

1.2.4 Symbols for certain types of information

Symbol Meaning
Permitted
Procedures, processes or actions that are permitted.
Preferred
Procedures, processes or actions that are preferred.
Forbidden
Procedures, processes or actions that are forbidden.
Tip
Indicates additional information.
Reference to documentation
Reference to page
Reference to graphic
Notice or individual step to be observed
, 2., 3.… Series of steps
Result of a step
Help in the event of a problem
Visual inspection

1.2.5 Symbols in graphics

Symbol Meaning
1, 2, 3,... Item numbers
, 2., 3.… Series of steps
A, B, C, ... Views
A-A, B-B, C-C, ... Sections
Hazardous area
Safe area (non-hazardous area)
Flow direction
Endress+Hauser 7
Page 8
Document information Proline Promass F 200 HART

1.3 Documentation

For an overview of the scope of the associated Technical Documentation, refer to the following:
• The W@M Device Viewer : Enter the serial number from the nameplate (www.endress.com/deviceviewer)
• The Endress+Hauser Operations App: Enter the serial number from the nameplate or scan the 2-D matrix code (QR code) on the nameplate.
For a detailed list of the individual documents along with the documentation code →  148

1.3.1 Standard documentation

Document type Purpose and content of the document
Technical Information Planning aid for your device
The document contains all the technical data on the device and provides an overview of the accessories and other products that can be ordered for the device.
Sensor Brief Operating Instructions Guides you quickly to the 1st measured value - Part 1
The Sensor Brief Operating Instructions are aimed at specialists with responsibility for installing the measuring device.
• Incoming acceptance and product identification
• Storage and transport
• Installation
Transmitter Brief Operating Instructions
Description of Device Parameters Reference for your parameters
Guides you quickly to the 1st measured value - Part 2
The Transmitter Brief Operating Instructions are aimed at specialists with responsibility for commissioning, configuring and parameterizing the measuring device (until the first measured value).
• Product description
• Installation
• Electrical connection
• Operation options
• System integration
• Commissioning
• Diagnostic information
The document provides a detailed explanation of each individual parameter in the Expert operating menu. The description is aimed at those who work with the device over the entire life cycle and perform specific configurations.

1.3.2 Supplementary device-dependent documentation

Additional documents are supplied depending on the device version ordered: Always comply strictly with the instructions in the supplementary documentation. The supplementary documentation is an integral part of the device documentation.

1.4 Registered trademarks

®
HART
Registered trademark of the FieldComm Group, Austin, Texas, USA
TRI-CLAMP
Registered trademark of Ladish & Co., Inc., Kenosha, USA
Applicator®, FieldCare®, DeviceCare ®, Field XpertTM, HistoROM®, Heartbeat Technology
Registered or registration-pending trademarks of the Endress+Hauser Group
8 Endress+Hauser
®
TM
Page 9
Proline Promass F 200 HART Basic safety instructions

2 Basic safety instructions

2.1 Requirements for the personnel

The personnel for installation, commissioning, diagnostics and maintenance must fulfill the following requirements:
Trained, qualified specialists must have a relevant qualification for this specific function
and task. Are authorized by the plant owner/operator.
Are familiar with federal/national regulations.
Before starting work, read and understand the instructions in the manual and
supplementary documentation as well as the certificates (depending on the application). Follow instructions and comply with basic conditions.
The operating personnel must fulfill the following requirements:
Are instructed and authorized according to the requirements of the task by the facility's
owner-operator. Follow the instructions in this manual.

2.2 Designated use

Application and media
The measuring device described in these Instructions is intended only for flow measurement of liquids and gases.
Depending on the version ordered, the measuring device can also measure potentially explosive, flammable, poisonous and oxidizing media.
Measuring devices for use in hazardous areas, in hygienic applications or where there is an increased risk due to process pressure, are labeled accordingly on the nameplate.
To ensure that the measuring device remains in proper condition for the operation time:
Only use the measuring device in full compliance with the data on the nameplate and
the general conditions listed in the Operating Instructions and supplementary documentation. Based on the nameplate, check whether the ordered device is permitted for the
intended use in the hazardous area (e.g. explosion protection, pressure vessel safety). Use the measuring device only for media to which the process-wetted materials are
sufficiently resistant. If the measuring device is not operated at atmospheric temperature, compliance with
the relevant basic conditions specified in the associated device documentation is absolutely essential: "Documentation" section.→  8. Protect the measuring device permanently against corrosion from environmental
influences.
Incorrect use
Non-designated use can compromise safety. The manufacturer is not liable for damage caused by improper or non-designated use.
WARNING
L
Danger of breakage due to corrosive or abrasive fluids!
Verify the compatibility of the process fluid with the sensor material.
Ensure the resistance of all fluid-wetted materials in the process.
Keep within the specified pressure and temperature range.
Endress+Hauser 9
Page 10
Basic safety instructions Proline Promass F 200 HART
NOTICE
Verification for borderline cases:
For special fluids and fluids for cleaning, Endress+Hauser is glad to provide assistance
in verifying the corrosion resistance of fluid-wetted materials, but does not accept any warranty or liability as minute changes in the temperature, concentration or level of contamination in the process can alter the corrosion resistance properties.
Residual risks
WARNING
L
The electronics and the medium may cause the surfaces to heat up. This presents a burn hazard!
For elevated fluid temperatures, ensure protection against contact to prevent burns.
WARNING
L
Danger of housing breaking due to measuring tube breakage!
In the event of a measuring tube breakage for a device version without rupture disk it is
possible for the pressure loading capacity of the sensor housing to be exceeded. This can lead to rupture or failure of the sensor housing.

2.3 Workplace safety

For work on and with the device:
Wear the required personal protective equipment according to federal/national
regulations.
For welding work on the piping:
Do not ground the welding unit via the measuring device.
If working on and with the device with wet hands:
Due to the increased risk of electric shock, gloves must be worn.

2.4 Operational safety

Risk of injury.
Operate the device in proper technical condition and fail-safe condition only.
The operator is responsible for interference-free operation of the device.
Conversions to the device
Unauthorized modifications to the device are not permitted and can lead to unforeseeable dangers.
If, despite this, modifications are required, consult with Endress+Hauser.
Repair
To ensure continued operational safety and reliability,
Carry out repairs on the device only if they are expressly permitted.
Observe federal/national regulations pertaining to repair of an electrical device.
Use original spare parts and accessories from Endress+Hauser only.

2.5 Product safety

This measuring device is designed in accordance with good engineering practice to meet state-of-the-art safety requirements, has been tested, and left the factory in a condition in which it is safe to operate.
10 Endress+Hauser
Page 11
Proline Promass F 200 HART Basic safety instructions
It meets general safety standards and legal requirements. It also complies with the EU directives listed in the device-specific EU Declaration of Conformity. Endress+Hauser confirms this by affixing the CE mark to the device.

2.6 IT security

We only provide a warranty if the device is installed and used as described in the Operating Instructions. The device is equipped with security mechanisms to protect it against any inadvertent changes to the device settings.
IT security measures in line with operators' security standards and designed to provide additional protection for the device and device data transfer must be implemented by the operators themselves.

2.7 Device-specific IT security

The device offers a range of specific functions to support protective measures on the operator's side. These functions can be configured by the user and guarantee greater in­operation safety if used correctly. An overview of the most important functions is provided in the following section.

2.7.1 Protecting access via hardware write protection

Write access to the device parameters via the local display or operating tool (e.g. FieldCare, DeviceCare) can be disabled via a write protection switch (DIP switch on the motherboard). When hardware write protection is enabled, only read access to the parameters is possible.

2.7.2 Protecting access via a password

A password can be used to protect against write access to the device parameters. This password locks write access to the device parameters via the local display or another
operating tool (e.g. FieldCare, DeviceCare) and, in terms of functionality, is equivalent to hardware write protection. If the service interface CDI RJ-45 is used, read access is only possible if the password is entered.
User-specific access code
Write access to the device parameters via the local display or operating tool (e.g. FieldCare, DeviceCare) can be protected by the modifiable, user-specific access code (→  91).
When the device is delivered, the device does not have an access code and is equivalent to 0000 (open).
General notes on the use of passwords
• The access code and network key supplied with the device should be changed during commissioning.
• Follow the general rules for generating a secure password when defining and managing the access code or network key.
• The user is responsible for the management and careful handling of the access code and network key.
• For information on configuring the access code or on what to do if you lose the password, see the "Write protection via access code" section →  91

2.7.3 Access via fieldbus

Cyclic fieldbus communication (read and write, e.g. measured value transmission) with a higher-order system is not affected by the restrictions mentioned above.
Endress+Hauser 11
Page 12
Product description Proline Promass F 200 HART
1 2
3
4
5 6
7
8
9
+
E

3 Product description

The device consists of a transmitter and a sensor. The device is available as a compact version:
The transmitter and sensor form a mechanical unit.

3.1 Product design

A0014056
 1 Important components of a measuring device
1 Electronics compartment cover 2 Display module 3 Main electronics module 4 Cable glands 5 Transmitter housing (incl. integrated HistoROM) 6 I/O electronics module 7 Terminals (pluggable spring terminals) 8 Connection compartment cover 9 Sensor
12 Endress+Hauser
Page 13

Proline Promass F 200 HART Incoming acceptance and product identification

1 2
1 2
Order code: Ser. no.: Ext. ord. cd.:
i
i
Date:
4 Incoming acceptance and product
identification

4.1 Incoming acceptance

Are the order codes on the
A0028673
delivery note (1) and the product sticker (2) identical?
Are the goods undamaged?
A0028673
Do the nameplate data
A0028673
A0028673
match the ordering information on the delivery note?
Is the CD-ROM with the Technical Documentation (depends on device version) and documents present?
• If one of the conditions is not satisfied, contact your Endress+Hauser Sales Center.
• Depending on the device version, the CD-ROM might not be part of the delivery! The Technical Documentation is available via the Internet or via the Endress+Hauser Operations App, see the "Product identification" section →  14.

4.2 Product identification

The following options are available for identification of the measuring device:
• Nameplate specifications
• Order code with breakdown of the device features on the delivery note
• Enter serial numbers from nameplates in W@M Device Viewer (www.endress.com/deviceviewer): All information about the measuring device is displayed.
• Enter the serial number from the nameplates into the Endress+Hauser Operations App or scan the 2-D matrix code (QR code) on the nameplate with the Endress+Hauser
Endress+Hauser 13
Operations App: all the information for the measuring device is displayed.
Page 14
Incoming acceptance and product identification Proline Promass F 200 HART
Order code:
Ext. ord. cd.:
Ser. no.:
Date:
i
i
Patents
322540-0001
1 2
3 4 5
6
7 8
9
10 11
12
14
15
16
17
13
For an overview of the scope of the associated Technical Documentation, refer to the following:
• The chapters "Additional standard documentation on the device" →  8 and "Supplementary device-dependent documentation" →  8
• The W@M Device Viewer: Enter the serial number from the nameplate (www.endress.com/deviceviewer)
• The Endress+Hauser Operations App: Enter the serial number from the nameplate or scan the 2-D matrix code (QR code) on the nameplate.

4.2.1 Transmitter nameplate

A0032237
 2 Example of a transmitter nameplate
1 Manufacturing location 2 Name of the transmitter 3 Order code 4 Serial number (ser. no.) 5 Extended order code (Ext. ord. cd.) 6 Electrical connection data, e.g. available inputs and outputs, supply voltage 7 Type of cable glands 8 Permitted ambient temperature (Ta)
9 Firmware version (FW) and device revision (Dev.Rev.) from the factory 10 CE mark, C-Tick 11 Additional information on version: certificates, approvals 12 Permitted temperature range for cable 13 Manufacturing date: year-month 14 Degree of protection 15 Approval information for explosion protection 16 Document number of safety-related supplementary documentation →  148 17 2-D matrix code
14 Endress+Hauser
Page 15
Proline Promass F 200 HART Incoming acceptance and product identification
Order code: Ser. no.: Ext. ord. cd.:
i
i
Date:
1
3 4 5
6
7
8
9
2
10
1112

4.2.2 Sensor nameplate

A0029199
 3 Example of a sensor nameplate
1 Name of the sensor 2 Manufacturing location 3 Order code 4 Serial number (ser. no.) 5 Extended order code (Ext. ord. cd.) 6 Nominal diameter of the sensor; flange nominal diameter/nominal pressure; sensor test pressure; medium
temperature range; material of measuring tube and manifold; sensor-specific information: e.g. pressure range
of secondary containment, wide-range density specification (special density calibration) 7 Approval information for explosion protection, Pressure Equipment Directive and degree of protection 8 Flow direction 9 Manufacturing date: year-month 10 2-D matrix code 11 Document number of safety-related supplementary documentation 12 CE mark, C-Tick 13 Surface roughness 14 Permitted ambient temperature (Ta)
Order code
The measuring device is reordered using the order code.
Extended order code
• The device type (product root) and basic specifications (mandatory features) are always listed.
• Of the optional specifications (optional features), only the safety and approval­related specifications are listed (e.g. LA). If other optional specifications are also ordered, these are indicated collectively using the # placeholder symbol (e.g. #LA#).
• If the ordered optional specifications do not include any safety and approval-related specifications, they are indicated by the + placeholder symbol (e.g. XXXXXX-ABCDE +).
Endress+Hauser 15
Page 16
Incoming acceptance and product identification Proline Promass F 200 HART

4.2.3 Symbols on measuring device

Symbol Meaning
WARNING!
This symbol alerts you to a dangerous situation. Failure to avoid this situation can result in serious or fatal injury.
Reference to documentation
Refers to the corresponding device documentation.
Protective ground connection
A terminal which must be connected to ground prior to establishing any other connections.
16 Endress+Hauser
Page 17
Proline Promass F 200 HART Storage and transport

5 Storage and transport

5.1 Storage conditions

Observe the following notes for storage:
• Store in the original packaging to ensure protection from shock.
• Do not remove protective covers or protective caps installed on process connections. They prevent mechanical damage to the sealing surfaces and contamination in the measuring tube.
• Protect from direct sunlight to avoid unacceptably high surface temperatures.
• Store in a dry and dust-free place.
• Do not store outdoors.
Storage temperature→  138

5.2 Transporting the product

Transport the measuring device to the measuring point in the original packaging.
A0029252
Do not remove protective covers or caps installed on process connections. They prevent mechanical damage to the sealing surfaces and contamination in the measuring tube.

5.2.1 Measuring devices without lifting lugs

WARNING
L
Center of gravity of the measuring device is higher than the suspension points of the webbing slings.
Risk of injury if the measuring device slips.
Secure the measuring device against slipping or turning.
Observe the weight specified on the packaging (stick-on label).
A0029214
Endress+Hauser 17
Page 18
Storage and transport Proline Promass F 200 HART

5.2.2 Measuring devices with lifting lugs

CAUTION
L
Special transportation instructions for devices with lifting lugs
Only use the lifting lugs fitted on the device or flanges to transport the device.
The device must always be secured at two lifting lugs at least.

5.2.3 Transporting with a fork lift

If transporting in wood crates, the floor structure enables the crates to be lifted lengthwise or at both sides using a forklift.

5.3 Packaging disposal

All packaging materials are environmentally friendly and 100% recyclable:
• Measuring device secondary packaging: polymer stretch film that conforms to EC Directive 2002/95/EC (RoHS).
• Packaging: – Wood crate, treated in accordance with ISPM 15 standard, which is confirmed by the
affixed IPPC logo. or
– Carton in accordance with European Packaging Directive 94/62EC; recyclability is
confirmed by the affixed RESY symbol.
• Seaworthy packaging (optional): Wood crate, treated in accordance with ISPM 15 standard, which is confirmed by the affixed IPPC logo.
• Carrying and mounting hardware: – Disposable plastic pallet – Plastic straps – Plastic adhesive strips
• Dunnage: Paper cushion
18 Endress+Hauser
Page 19
Proline Promass F 200 HART Installation
1
2
3
4
5

6 Installation

6.1 Installation conditions

No special measures such as supports are necessary. External forces are absorbed by the construction of the device.

6.1.1 Mounting position

Mounting location
To prevent measuring errors arising from accumulation of gas bubbles in the measuring tube, avoid the following mounting locations in the pipe:
• Highest point of a pipeline.
• Directly upstream of a free pipe outlet in a down pipe.
Installation in down pipes
However, the following installation suggestion allows for installation in an open vertical pipeline. Pipe restrictions or the use of an orifice with a smaller cross-section than the nominal diameter prevent the sensor running empty while measurement is in progress.
A0028772
A0028773
 4 Installation in a down pipe (e.g. for batching applications)
1 Supply tank 2 Sensor 3 Orifice plate, pipe restriction 4 Valve 5 Batching tank
Endress+Hauser 19
Page 20
Installation Proline Promass F 200 HART
1 2
DN Ø orifice plate, pipe restriction
[mm] [in] [mm] [in]
8 ³⁄₈ 6 0.24
15 ½ 10 0.40
25 1 14 0.55
40 22 0.87
50 2 28 1.10
80 3 50 1.97
Orientation
The direction of the arrow on the sensor nameplate helps you to install the sensor according to the flow direction (direction of medium flow through the piping).
Orientation Recommendation
A Vertical orientation
A0015591
B Horizontal orientation, transmitter at
top
1)
Exceptions:
→  5,  20
A0015589
C Horizontal orientation, transmitter at
bottom
2)
Exceptions:
→  5,  20
A0015590
D Horizontal orientation, transmitter at
side
A0015592
1) Applications with low process temperatures may decrease the ambient temperature. To maintain the
minimum ambient temperature for the transmitter, this orientation is recommended.
2) Applications with high process temperatures may increase the ambient temperature. To maintain the
maximum ambient temperature for the transmitter, this orientation is recommended.
If a sensor is installed horizontally with a curved measuring tube, match the position of the sensor to the fluid properties.
A0028774
 5 Orientation of sensor with curved measuring tube
1 Avoid this orientation for fluids with entrained solids: Risk of solids accumulating. 2 Avoid this orientation for outgassing fluids: Risk of gas accumulating.
20 Endress+Hauser
Page 21
Proline Promass F 200 HART Installation
Inlet and outlet runs
No special precautions need to be taken for fittings which create turbulence, such as valves, elbows or T-pieces, as long as no cavitation occurs →  21.
A0029322 A0029323
Installation dimensions
For the dimensions and installation lengths of the device, see the "Technical Information" document, "Mechanical construction" section.

6.1.2 Requirements from environment and process

Ambient temperature range
Measuring device –40 to +60 °C (–40 to +140 °F)
Readability of the local display –20 to +60 °C (–4 to +140 °F)
The readability of the display may be impaired at temperatures outside the temperature range.
If operating outdoors:
Avoid direct sunlight, particularly in warm climatic regions.
You can order a weather protection cover from Endress+Hauser : →  123
System pressure
It is important that cavitation does not occur, or that gases entrained in the liquids do not outgas.
Cavitation is caused if the pressure drops below the vapor pressure:
• In liquids that have a low boiling point (e.g. hydrocarbons, solvents, liquefied gases)
• In suction lines Ensure the system pressure is sufficiently high to prevent cavitation and outgassing.
For this reason, the following mounting locations are recommended:
• At the lowest point in a vertical pipe
• Downstream from pumps (no danger of vacuum)
A0028777
Thermal insulation
In the case of some fluids, it is important that the heat radiated from the sensor to the transmitter is kept to a minimum. A wide range of materials can be used for the required insulation.
Endress+Hauser 21
Page 22
Installation Proline Promass F 200 HART
t
a
0
10
30
20
40
[ ]mm[ ]in
80 90 100 110
[°C]
[°F]
200 250 290
0.5
1.0
1.5
0
t
40(104)
t
60(140)
t
120 130 140
T
m
150
NOTICE
Electronics overheating on account of thermal insulation!
Observe maximum permitted insulation height of the transmitter neck so that the
transmitter head is completely free.
NOTICE
Danger of overheating with insulation
Ensure that the temperature at the lower end of the transmitter housing does not
exceed 80 °C (176 °F)
NOTICE
The insulation can also be thicker than the maximum recommended insulation thickness.
Prerequisite:
Ensure that convection takes place on a sufficiently large scale at the transmitter neck.
Ensure that a sufficiently large area of the housing support remains exposed. The
uncovered part serves as a radiator and protects the electronics from overheating and excessive cooling.
t Maximum insulation thickness a Minimum distance to insulation
The minimum distance a between the transmitter and the insulation is 20 mm (0.79 in). This is to ensure that the transmitter remains completely exposed.
Maximum recommended insulation thickness
 6 Maximum recommended insulation thickness depending on the temperature of the medium and the
ambient temperature
t Insulation thickness
T
m
t40
(104)
t60
(140)
22 Endress+Hauser
Medium temperature
Maximum recommended insulation thickness at an ambient temperature of Ta = 40 °C (104 °F)
Maximum recommended insulation thickness at an ambient temperature of Ta = 60 °C (140 °F)
A0028853
A0028904
Page 23
Proline Promass F 200 HART Installation
0
20
60
40
80
1
2
3
0
t
100
4
[ ]mm
t
40(104)
t
60(140)
80 100 120 140
200 300 400
160 180 200
[°F]
T
m
[°C]
Maximum recommended insulation thickness for the extended temperature range or insulation
For the extended temperature range, version with long extension neck, order code for "Measuring tube material", option SD, SE, SF, TH or extension neck for insulation, order code for "Sensor option", option CG:
A0029921
t Insulation thickness
T
t40
t60
m
(104)
(140)
Medium temperature
Maximum recommended insulation thickness at an ambient temperature of Ta = 40 °C (104 °F)
Maximum recommended insulation thickness at an ambient temperature of Ta = 60 °C (140 °F)
Heating
NOTICE
Electronics can overheat due to elevated ambient temperature!
Observe maximum permitted ambient temperature for the transmitter .
Depending on the fluid temperature, take the device orientation requirements into
account .
NOTICE
Danger of overheating when heating
Ensure that the temperature at the lower end of the transmitter housing does not
exceed 80 °C (176 °F). Ensure that convection takes place on a sufficiently large scale at the transmitter neck.
Ensure that a sufficiently large area of the housing support remains exposed. The
uncovered part serves as a radiator and protects the electronics from overheating and excessive cooling.
Heating options
If a fluid requires that no heat loss should occur at the sensor, users can avail of the following heating options:
• Electrical heating, e.g. with electric band heaters
• Via pipes carrying hot water or steam
• Via heating jackets
Using an electrical trace heating system
Endress+Hauser 23
If heating is regulated via phase angle control or pulse packages, magnetic fields can affect the measured values (= for values that are greater than the values permitted by the EN standard (sine 30 A/m)).
For this reason, the sensor must be magnetically shielded: the housing can be shielded with tin plates or electric sheets without a privileged direction (e.g. V330-35A).
Page 24
Installation Proline Promass F 200 HART
The sheet must have the following properties:
• Relative magnetic permeability µr ≥ 300
• Plate thickness d ≥ 0.35 mm (d ≥ 0.014 in)
Vibrations
The high oscillation frequency of the measuring tubes ensures that the correct operation of the measuring system is not influenced by plant vibrations.

6.1.3 Special mounting instructions

Rupture disk
Information that is relevant to the process: →  140.
WARNING
L
Limited functional reliability of the rupture disk.
Danger to persons from escaping fluids!
Do not remove the rupture disk.
When using a rupture disk, do not use a heating jacket.
Make sure that the function and operation of the rupture disk is not impeded through
the installation of the device. Take precautions to prevent damage and danger to persons if the rupture disk is
actuated. Observe information on the rupture disk sticker.
The position of the rupture disk is indicated on a sticker beside it. The transportation guard must be removed.
For information on the dimensions: see the "Mechanical construction" section
The existing connecting nozzles are not intended for the purpose of rinsing or pressure monitoring, but instead serve as the mounting location for the rupture disk.
In the event of a failure of the rupture disk, a discharge device can be screwed onto the internal thread of the rupture disk in order to drain off any escaping medium.
Zero point adjustment
All measuring devices are calibrated in accordance with state-of-the-art technology. Calibration takes place under reference conditions→  133. Therefore, a zero point adjustment in the field is generally not required.
Experience shows that zero point adjustment is advisable only in special cases:
• To achieve maximum measuring accuracy even with low flow rates.
• Under extreme process or operating conditions (e.g. very high process temperatures or very high-viscosity fluids).

6.2 Mounting the measuring device

6.2.1 Required tools

For transmitter
• For turning the transmitter housing: Open-ended wrench8 mm
• For opening the securing clamps: Allen key3 mm
• For turning the transmitter housing: Open-ended wrench8 mm
• For opening the securing clamps: Allen key3 mm
24 Endress+Hauser
Page 25
Proline Promass F 200 HART Installation
max. 350°
8 mm
8 mm
1.
2.
3.
For sensor
For flanges and other process connections: Corresponding mounting tools

6.2.2 Preparing the measuring device

1. Remove all remaining transport packaging.
2. Remove any protective covers or protective caps present from the sensor.
3. Remove stick-on label on the electronics compartment cover.

6.2.3 Mounting the measuring device

WARNING
L
Danger due to improper process sealing!
Ensure that the inside diameters of the gaskets are greater than or equal to that of the
process connections and piping. Ensure that the gaskets are clean and undamaged.
Install the gaskets correctly.
1. Ensure that the direction of the arrow on the nameplate of the sensor matches the flow direction of the fluid.
2. Install the measuring device or turn the transmitter housing so that the cable entries do not point upwards.
A0029263

6.2.4 Turning the transmitter housing

To provide easier access to the connection compartment or display module, the transmitter housing can be turned.
A0032242
1. Release the fixing screw.
2. Turn the housing to the desired position.
3. Firmly tighten the securing screw.
Endress+Hauser 25
Page 26
Installation Proline Promass F 200 HART
+
E
1
3 mm
1.
2.
3.
4.

6.2.5 Turning the display module

The display module can be turned to optimize display readability and operability.
A0032238
1. Loosen the securing clamp of the electronics compartment cover using an Allen key.
2. Unscrew cover of the electronics compartment from the transmitter housing.
3. Optional: pull out the display module with a gentle rotational movement.
4. Turn the display module to the desired position: max. 8 × 45° in every direction.
5. Without display module pulled out: Allow display module to engage at desired position.
6. With display module pulled out: Feed the cable into the gap between the housing and main electronics module and
plug the display module into the electronics compartment until it engages.
7. Reverse the removal procedure to reassemble the transmitter.

6.3 Post-installation check

Is the device undamaged (visual inspection)?
Does the measuring device conform to the measuring point specifications?
For example:
• Process temperature →  138
• Process pressure (refer to the chapter on "Pressure-temperature ratings" of the "Technical Information" document)
• Ambient temperature
• Measuring range
Has the correct orientation for the sensor been selected ?
• According to sensor type
• According to medium temperature
• According to medium properties (outgassing, with entrained solids)
Does the arrow on the sensor nameplate match the direction of flow of the fluid through the piping →  20?
Are the measuring point identification and labeling correct (visual inspection)?
Is the device adequately protected from precipitation and direct sunlight?
Are the securing screw and securing clamp tightened securely?
26 Endress+Hauser
Page 27
Proline Promass F 200 HART Electrical connection

7 Electrical connection

The measuring device does not have an internal circuit breaker. For this reason, assign the measuring device a switch or power-circuit breaker so that the power supply line can be easily disconnected from the mains.

7.1 Connection conditions

7.1.1 Required tools

• For cable entries: Use corresponding tools
• For securing clamp: Allen key 3 mm
• Wire stripper
• When using stranded cables: Crimper for wire end ferrule
• For removing cables from terminal: Flat blade screwdriver ≤ 3 mm (0.12 in)

7.1.2 Connecting cable requirements

The connecting cables provided by the customer must fulfill the following requirements.
Electrical safety
In accordance with applicable federal/national regulations.
Permitted temperature range
Minimum requirement: cable temperature range ≥ ambient temperature +20 K
Signal cable
Current output 4 to 20 mA HART
A shielded cable is recommended. Observe grounding concept of the plant.
Current output 4 to 20 mA
Standard installation cable is sufficient.
Pulse/frequency/switch output
Standard installation cable is sufficient.
Cable diameter
• Cable glands supplied: M20 × 1.5 with cable  6 to 12 mm (0.24 to 0.47 in)
• Plug-in spring terminals for device version without integrated overvoltage protection: wire cross-sections 0.5 to 2.5 mm2 (20 to 14 AWG)
• Screw terminals for device version with integrated overvoltage protection: wire cross­sections 0.2 to 2.5 mm2 (24 to 14 AWG)
Endress+Hauser 27
Page 28
Electrical connection Proline Promass F 200 HART
+ 1
– 2
– 4
+ 3
1
2
3
+ 1
– 2
4
+
3
1
2
3

7.1.3 Terminal assignment

Transmitter
4-20 mA HART connection version with additional outputs
A0013570
Maximum number of terminals, without integrated overvoltage protection
1
Output 1 (passive): supply voltage and signal transmission
2
Output 2 (passive): supply voltage and signal transmission
3
Ground terminal for cable shield
Order code for "Output" Terminal numbers
Output 1 Output 2
1 (+) 2 (-) 3 (+) 4 (-)
Option A 4-20 mA HART (passive) -
Option C
1)
1)
4-20 mA HART (passive)
4-20 mA HART (passive) 4-20 mA analog (passive)
Option B
1) Output 1 must always be used; output 2 is optional.
Maximum number of terminals, with integrated overvoltage protection

7.1.4 Requirements for the supply unit

Supply voltage
A0018161
Pulse/frequency/switch output
(passive)
Transmitter
An external power supply is required for each output. The following supply voltage values apply for the outputs available:
Order code for "Output"
Option A
Option B switch output
Option C
1) External supply voltage of the power supply unit with load.
2) For device versions with SD03 local display: The terminal voltage must be increased by DC 2 V if
28 Endress+Hauser
1) 2)
: 4-20 mA HART • For 4 mA: ≥ DC 17.9 V
1) 2)
: 4-20 mA HART, pulse/frequency/
1) 2)
: 4-20 mA HART + 4-20 mA analog • For 4 mA: ≥ DC 17.9 V
backlighting is used.
Minimum
terminal voltage
• For 20 mA: ≥ DC 13.5 V
• For 4 mA: ≥ DC 17.9 V
• For 20 mA: ≥ DC 13.5 V
• For 20 mA: ≥ DC 13.5 V
Maximum
terminal voltage
DC 35 V
DC 35 V
DC 30 V
Page 29
Proline Promass F 200 HART Electrical connection
0
100
200
300
400
500
16 18 20
22 24
26 28 30 32
U [V]
S
RB[Ω]
34 36
A B
Load
Load for current output: 0 to 500 Ω, depending on the external supply voltage of the power supply unit
Calculation of the maximum load
Depending on the supply voltage of the power supply unit (US), the maximum load (RB) including line resistance must be observed to ensure adequate terminal voltage at the device. In doing so, observe the minimum terminal voltage
• For US = 17.9 to 18.9 V: RB ≤ (US - 17.9 V): 0.0036 A
• For US = 18.9 to 24 V: RB ≤ (US - 13 V): 0.022 A
• For US = ≥ 24 V: RB ≤ 500 Ω
A Operating range for order code for "Output", option A "4-20 mA HART"/option B "4-20 mA HART, pulse/
frequency/switch output" with Ex i and option C "4-20 mA HART + 4-20 mA analog"
B Operating range for order code for "Output", option A "4-20 mA HART"/option B "4-20 mA HART, pulse/
frequency/switch output" with non-Ex and Ex d
Sample calculation
Supply voltage of the power supply unit: US = 19 V Maximum load: RB ≤ (19 V - 13 V): 0.022 A = 273 Ω

7.1.5 Preparing the measuring device

NOTICE
Insufficient sealing of the housing!
Operational reliability of the measuring device could be compromised.
Use suitable cable glands corresponding to the degree of protection.
1. Remove dummy plug if present.
2. If the measuring device is supplied without cable glands: Provide suitable cable gland for corresponding connecting cable.
3. If the measuring device is supplied with cable glands: Observe requirements for connecting cables →  27.
A0013563
Endress+Hauser 29
Page 30
Electrical connection Proline Promass F 200 HART
10 (0.4)
mm (in)
20 mm3 mm
1.2. 3.
4.
5. 6.

7.2 Connecting the measuring device

NOTICE
Limitation of electrical safety due to incorrect connection!
Have electrical connection work carried out by correspondingly trained specialists only.
Observe applicable federal/national installation codes and regulations.
Comply with local workplace safety regulations.
Always connect the protective ground cable before connecting additional cables.
For use in potentially explosive atmospheres, observe the information in the device-
specific Ex documentation.

7.2.1 Connecting the transmitter

Connection via terminals
1. Loosen the securing clamp of the connection compartment cover.
2. Unscrew the connection compartment cover.
3. Push the cable through the cable entry . To ensure tight sealing, do not remove the sealing ring from the cable entry.
4. Strip the cable and cable ends. In the case of stranded cables, also fit ferrules.
5. Connect cable in accordance with terminal assignment→  28. For HART communication: when connecting the cable shielding to the ground clamp, observe the grounding concept of the facility.
6. Firmly tighten the cable glands.
7.
WARNING
L
Housing degree of protection may be voided due to insufficient sealing of the housing.
Screw in the screw without using any lubricant. The threads on the cover are
coated with a dry lubricant.
Reverse the removal procedure to reassemble the transmitter.
A0032239
30 Endress+Hauser
Page 31
Proline Promass F 200 HART Electrical connection
213 4
3 (0.12)
2
3
4...20 mA
41
5
Removing a cable
A0032240
To remove a cable from the terminal, use a flat-blade screwdriver to push the slot
between the two terminal holes while simultaneously pulling the cable end out of the terminal.

7.2.2 Ensuring potential equalization

Requirements
No special measures for potential equalization are required.
For devices intended for use in hazardous locations, please observe the guidelines in the Ex documentation (XA).

7.3 Special connection instructions

7.3.1 Connection examples

Current output 4-20 mA HART
A0028762
 7 Connection example for 4 to 20 mA HART current output (passive)
1 Automation system with current input (e.g. PLC) 2 Power supply 3 Cable shield: the cable shield must be grounded at both ends to comply with EMC requirements; observe cable
specifications 4 Analog display unit: observe maximum load →  29 5 Transmitter
Endress+Hauser 31
Page 32
Electrical connection Proline Promass F 200 HART
1
2
3
12345
1
2
3
Pulse/frequency output
A0028761
 8 Connection example for pulse/frequency output (passive)
1 Automation system with pulse/frequency input (e.g. PLC) 2 Power supply 3 Transmitter: Observe input values →  130
Switch output
 9 Connection example for switch output (passive)
1 Automation system with switch input (e.g. PLC) 2 Power supply 3 Transmitter: Observe input values →  130
A0028760
32 Endress+Hauser
Page 33
Proline Promass F 200 HART Electrical connection
2
4...20 mA
4
1
2
3
3
6
5
HART input
A0028763
 10 Connection example for HART input with a common negative (passive)
1 Automation system with HART output (e.g. PLC) 2 Active barrier for power supply (e.g. RN221N) 3 Cable shield: the cable shield must be grounded at both ends to comply with EMC requirements; observe cable
specifications 4 Analog display unit: observe maximum load →  29 5 Pressure measuring device (e.g. Cerabar M, Cerabar S): see requirements 6 Transmitter

7.4 Ensuring the degree of protection

The measuring device fulfills all the requirements for the IP66/67 degree of protection, Type 4X enclosure.
To guarantee IP66/67 degree of protection, Type 4X enclosure, carry out the following steps after the electrical connection:
1. Check that the housing seals are clean and fitted correctly.
2. Dry, clean or replace the seals if necessary.
3. Tighten all housing screws and screw covers.
4. Firmly tighten the cable glands.
5. To ensure that moisture does not enter the cable entry: Route the cable so that it loops down before the cable entry ("water trap"). 
6. Insert dummy plugs into unused cable entries.
A0029278
Endress+Hauser 33
Page 34
Electrical connection Proline Promass F 200 HART

7.5 Post-connection check

Are cables or the device undamaged (visual inspection)?
Do the cables used meet the requirements→  27?
Do the cables have adequate strain relief?
Are all the cable glands installed, firmly tightened and leak-tight? Cable run with "water trap" →  33 ?
Depending on the device version: are all the device plugs firmly tightened →  30?
Does the supply voltage match the specifications on the transmitter nameplate →  28?
Is the terminal assignment correct →  28?
If supply voltage is present, do values appear on the display module?
Are all housing covers installed and firmly tightened?
Is the securing clamp tightened correctly?
34 Endress+Hauser
Page 35
Proline Promass F 200 HART Operation options
1
2 3 4
E+-
ESC
5
6

8 Operation options

8.1 Overview of operation options

A0032226
1 Local operation via display module 2 Computer with operating tool (e.g. FieldCare, DeviceCare, AMS Device Manager, SIMATIC PDM) 3 Field Xpert SFX350 or SFX370 4 Field Communicator 475 5 Control system (e.g. PLC) 6 VIATOR Bluetooth modem with connecting cable
Endress+Hauser 35
Page 36
Operation options Proline Promass F 200 HART
!
Expert
System
Sensor
Communication
Application
Diagnostics
Access status display
Output
Operating menu for experts
Language
Operatation Language
Parameter 1
Setup
Submenu 1
Submenu n
Device tag
Advanced setup
Enter access code
Parameter 1
Parameter n
Submenu 1
Submenu n
Diagnostics
Parameter 1
Parameter n
Submenu 1
Submenu n
Operating menu for operators and maintenances
Parameter n
Operator
Maintenance
Task-oriented
Function-oriented
Expert
Wizard 1 / Parameter 1
Wizard n / Parameter n
Parameter n
Input

8.2 Structure and function of the operating menu

8.2.1 Structure of the operating menu

For an overview of the operating menu for experts: "Description of Device Parameters" document supplied with the device
36 Endress+Hauser
 11 Schematic structure of the operating menu
A0018237-EN
Page 37
Proline Promass F 200 HART Operation options

8.2.2 Operating philosophy

The individual parts of the operating menu are assigned to certain user roles (operator, maintenance etc.). Each user role contains typical tasks within the device lifecycle.
Menu/parameter User role and tasks Content/meaning
Language task-oriented Role "Operator", "Maintenance"
Tasks during operation:
Operation • Configuring the operational display (e.g. display format, display contrast)
Setup "Maintenance" role
Diagnostics "Maintenance" role
• Configuring the operational display
• Reading measured values
Commissioning:
• Configuration of the measurement
• Configuration of the inputs and outputs
Fault elimination:
• Diagnostics and elimination of process and device errors
• Measured value simulation
• Defining the operating language
• Resetting and controlling totalizers
• Resetting and controlling totalizers
Wizards for fast commissioning:
• Set the system units
• Define the medium
• Configure the outputs
• Configuring the operational display
• Define the output conditioning
• Set the low flow cut off
• Configure partial and empty pipe detection
Advanced setup
• For more customized configuration of the measurement (adaptation to special measuring conditions)
• Configuration of totalizers
• Configure the WLAN settings
• Administration (define access code, reset measuring device)
Contains all parameters for error detection and analyzing process and device errors:
• Diagnostic list Contains up to 5 currently pending diagnostic messages.
• Event logbook Contains event messages that have occurred.
• Device information Contains information for identifying the device.
• Measured values Contains all current measured values.
Data logging submenu with "Extended HistoROM" order option Storage and visualization of measured values
• Heartbeat The functionality of the device is checked on demand and the verification results are documented.
• Simulation Is used to simulate measured values or output values.
Expert function-oriented Tasks that require detailed
knowledge of the function of the device:
• Commissioning measurements under difficult conditions
• Optimal adaptation of the measurement to difficult conditions
• Detailed configuration of the communication interface
• Error diagnostics in difficult cases
Contains all the parameters of the device and makes it possible to access these parameters directly using an access code. The structure of this menu is based on the function blocks of the device:
• System Contains all higher-order device parameters which do not concern the measurement or the communication interface.
• Sensor Configuration of the measurement.
• Input Configuration of the input.
• Output Configuration of the outputs.
• Communication Configuration of the digital communication interface.
• Application Configure the functions that go beyond the actual measurement (e.g. totalizer).
• Diagnostics Error detection and analysis of process and device errors and for device simulation and Heartbeat Technology.
Endress+Hauser 37
Page 38
Operation options Proline Promass F 200 HART
X X X X X XXX X
2
1
3
kg/h
1120.50
F

8.3 Access to the operating menu via the local display

8.3.1 Operational display

A0029348
1 Operational display 2 Device tag 3 Status area 4 Display area for measured values (4-line) 5 Operating elements→  43
Status area
The following symbols appear in the status area of the operational display at the top right:
• Status signals→  103 – F: Failure – C: Function check – S: Out of specification – M: Maintenance required
• Diagnostic behavior→  104 – : Alarm – : Warning
: Locking (the device is locked via the hardware )
: Communication (communication via remote operation is active)
Display area
In the display area, each measured value is prefaced by certain symbol types for further description:
Measured variable Measurement channel
number
Example
Diagnostic behavior
Appears only if a diagnostics event is present for this measured variable.
Measured values
Symbol Meaning
Mass flow
• Volume flow
• Corrected volume flow
38 Endress+Hauser
Page 39
Proline Promass F 200 HART Operation options
4
2
1
3
/../Operation
0091-1
Access stat.disp
Operator
Locking status Display
4
2
1
3
/../Select medium
Select medium
Liquid
• Density
• Reference density
Temperature
Totalizer
The measurement channel number indicates which of the three totalizers is
displayed.
Output
The measurement channel number indicates which of the two current outputs is
displayed.
Measurement channel numbers
Symbol Meaning
Measurement channel 1 to 4
The measurement channel number is displayed only if more than one channel is present for the same measured variable type (e.g. Totalizer 1 to 3).
Diagnostic behavior
The diagnostic behavior pertains to a diagnostic event that is relevant to the displayed measured variable. For information on the symbols →  104
The number and display format of the measured values can be configured via the
Format display parameter (→  74).

8.3.2 Navigation view

In the submenu In the wizard
A0013993-EN
1
Navigation view
2
Navigation path to current position
3
Status area
4
Display area for navigation
5
Operating elements →  43
A0013995-EN
Navigation path
The navigation path - displayed at the top left in the navigation view - consists of the following elements:
• In the submenu: Display symbol for menu
Examples / ../ Display
• In the wizard: Display symbol for wizard
Omission symbol for operating menu levels in between
Name of current
• Submenu
• Wizard
• Parameters
Endress+Hauser 39
Page 40
Operation options Proline Promass F 200 HART
/ ../ Display
For more information about the icons in the menu, refer to the "Display area" section →  40
Status area
The following appears in the status area of the navigation view in the top right corner:
• In the submenu – The direct access code for the parameter you are navigating to (e.g. 0022-1) – If a diagnostic event is present, the diagnostic behavior and status signal
• In the wizard If a diagnostic event is present, the diagnostic behavior and status signal
• For information on the diagnostic behavior and status signal →  103
• For information on the function and entry of the direct access code →  45
Display area
Menus
Symbol Meaning
Operation
Appears:
• In the menu next to the "Operation" selection
• At the left in the navigation path in the Operation menu
Setup
Appears:
• In the menu next to the "Setup" selection
• At the left in the navigation path in the Setup menu
Diagnostics
Appears:
• In the menu next to the "Diagnostics" selection
• At the left in the navigation path in the Diagnostics menu
Expert
Appears:
• In the menu next to the "Expert" selection
• At the left in the navigation path in the Expert menu
Submenus, wizards, parameters
Symbol Meaning
Submenu
Wizard
Parameters within a wizard
No display symbol exists for parameters in submenus.
Locking
Symbol Meaning
Parameter locked
When displayed in front of a parameter name, indicates that the parameter is locked.
• By a user-specific access code
• By the hardware write protection switch
40 Endress+Hauser
Page 41
Proline Promass F 200 HART Operation options
3
2
1
4
3
4
0
1 2
9
5
6
8
7
20
ABC_
DEFG
User
HIJK
LMNO
PQRS
TUVW
XYZ
Aa1
3
2
1
4
0
9
.
Wizard operation
Symbol Meaning
Switches to the previous parameter.
Confirms the parameter value and switches to the next parameter.
Opens the editing view of the parameter.

8.3.3 Editing view

Numeric editor Text editor
A0013999
1
Editing view
2
Display area of the entered values
3
Input mask
4
Operating elements →  43
A0013941
Input mask
The following input symbols are available in the input mask of the numeric and text editor:
Numeric editor
Symbol Meaning
Selection of numbers from 0 to 9.
Inserts decimal separator at the input position.
Inserts minus sign at the input position.
Confirms selection.
Moves the input position one position to the left.
Exits the input without applying the changes.
Clears all entered characters.
Endress+Hauser 41
Page 42
Operation options Proline Promass F 200 HART
Aa1
XYZ
ABC
_
xyz
abc
_
~&
"'^
_
_
Text editor
Symbol Meaning
Toggle
• Between upper-case and lower-case letters
• For entering numbers
• For entering special characters
Selection of letters from A to Z.
Selection of letters from a to z.
Selection of special characters.
Confirms selection.
Switches to the selection of the correction tools.
Exits the input without applying the changes.
Clears all entered characters.
Correction symbols under
Symbol Meaning
Clears all entered characters.
Moves the input position one position to the right.
Moves the input position one position to the left.
Deletes one character immediately to the left of the input position.
42 Endress+Hauser
Page 43
Proline Promass F 200 HART Operation options
+
+
+
++

8.3.4 Operating elements

Key Meaning
Minus key
In a menu, submenu
Moves the selection bar upwards in a choose list.
With a Wizard
Confirms the parameter value and goes to the previous parameter.
With a text and numeric editor
In the input mask, moves the selection bar to the left (backwards).
Plus key
In a menu, submenu
Moves the selection bar downwards in a choose list.
With a Wizard
Confirms the parameter value and goes to the next parameter.
With a text and numeric editor
Moves the selection bar to the right (forwards) in an input screen.
Enter key
For operational display
• Pressing the key briefly opens the operating menu.
• Pressing the key for 2 s opens the context menu.
In a menu, submenu
• Pressing the key briefly: – Opens the selected menu, submenu or parameter. – Starts the wizard. – If help text is open, closes the help text of the parameter.
• Pressing the key for 2 s for parameter: If present, opens the help text for the function of the parameter.
With a Wizard
Opens the editing view of the parameter.
With a text and numeric editor
• Pressing the key briefly: – Opens the selected group. – Carries out the selected action.
• Pressing the key for 2 s confirms the edited parameter value.
Escape key combination (press keys simultaneously)
In a menu, submenu
• Pressing the key briefly: – Exits the current menu level and takes you to the next higher level. – If help text is open, closes the help text of the parameter.
• Pressing the key for 2 s returns you to the operational display ("home position").
With a Wizard
Exits the wizard and takes you to the next higher level.
With a text and numeric editor
Closes the text or numeric editor without applying changes.
Minus/Enter key combination (press the keys simultaneously)
Reduces the contrast (brighter setting).
Plus/Enter key combination (press and hold down the keys simultaneously)
Increases the contrast (darker setting).
Minus/Plus/Enter key combination (press the keys simultaneously)
For operational display
Enables or disables the keypad lock (only SD02 display module).

8.3.5 Opening the context menu

Using the context menu, the user can call up the following menus quickly and directly from the operational display:
Endress+Hauser 43
Page 44
Operation options Proline Promass F 200 HART
XXXXXXXXXX
l/h
20.50
Setup
Simulation
• Setup
• Configuration backup display
• Simulation
Calling up and closing the context menu
The user is in the operational display.
1. Press  for 2 s.  The context menu opens.
A0017421-EN
2. Press  +  simultaneously.  The context menu is closed and the operational display appears.
Calling up the menu via the context menu
1. Open the context menu.
2. Press  to navigate to the desired menu.
3. Press  to confirm the selection.  The selected menu opens.
44 Endress+Hauser
Page 45
Proline Promass F 200 HART Operation options
X X X X X X XX X
20.50
0104-1
2 s
0091-1
0098-1
0098-1
0098-1
X X X X X X XX X
10.50
2800
Hz
mA
3.
4.
5.
6.
7.
8.
1.
2.
Display/operat.
Display/operat. Setup
Main menu
English
Main menu
Format display
/ ../Display
Contrast display Display intervall
1 value, max.
Setup
Access stat.disp
/ ../Display/operat.
Display
Locking status
1 value, max.
/ ../Format display
2 values Val. large+2val.
Bargr. + 1 value
Locking status
/ ../Display/operat.
Display
1 value, max.
/ ../Format display
2 values
Val. large+2val.
Bargr. + 1 value
Operator
Language
Language

8.3.6 Navigating and selecting from list

Different operating elements are used to navigate through the operating menu. The navigation path is displayed on the left in the header. Icons are displayed in front of the individual menus. These icons are also shown in the header during navigation.
For an explanation of the navigation view with symbols and operating elements →  39
Example: Setting the number of displayed measured values to "2 values"
Endress+Hauser 45

8.3.7 Calling the parameter directly

A parameter number is assigned to every parameter to be able to access a parameter directly via the onsite display. Entering this access code in the Direct access parameter calls up the desired parameter directly.
Navigation path
Expert → Direct access
A0029562-EN
Page 46
Operation options Proline Promass F 200 HART
1
0914-2
Ent.accesscode Enteraccesscodetodisable writeprotec.
The direct access code consists of a 4-digit number and the channel number, which identifies the channel of a process variable: e.g. 0914-1. In the navigation view, this appears on the right-hand side in the header of the selected parameter.
A0029414
1 Direct access code
Note the following when entering the direct access code:
• The leading zeros in the direct access code do not have to be entered. Example: Input of "914" instead of "0914"
• If no channel number is entered, channel 1 is jumped to automatically. Example: Enter 0914Assign process variable parameter
• If a different channel is jumped to: Enter the direct access code with the corresponding channel number. Example: Enter 0914-2Assign process variable parameter
For the direct access codes of the individual parameters, see the "Description of Device Parameters" document for the device

8.3.8 Calling up help text

Help text is available for some parameters and can be called up from the navigation view. The help text provides a brief explanation of the parameter function and thereby supports swift and safe commissioning.
Calling up and closing the help text
The user is in the navigation view and the selection bar is on a parameter.
1. Press  for 2 s.  The help text for the selected parameter opens.
A0014002-EN
 12 Example: Help text for parameter "Enter access code"
2. Press  +  simultaneously.  The help text is closed.
46 Endress+Hauser
Page 47
Proline Promass F 200 HART Operation options
3x
001-FT-101
DEFG HIJK
LMNO
PQRS TUVW
XYZ
Aa1@
ABC
001-FT-101
DEFG HIJK
LMNO
PQRS TUVW
XYZ
Aa1@
ABC
001-FT-101
DEFG HIJK
ABC
001-FT-10
DEFG HIJK
ABC
001-FT-10
DEFG HIJK
ABC
1x
001-FT-10
DEFG HIJK
LMNO
PQRS TUVW
XYZ
Aa1@
ABC
001-FT-10
DEFG HIJK
LMNO
PQRS TUVW
XYZ
Aa1@
ABC
001-FT-10
DEFG HIJK
ABC
A a 1
@
001-FT-10
DEFG HIJK
ABC
A a 1
@
001-FT-10
3456 789
= + - *
/
[ ]
( )
< >
{ }
Aa1@
012
001-FT-10
3456 789
012
0 1 2
001-FT-10
3456 789
012
0 1 2
001-FT-102
3456 789
012
0 1 2
001-FT-102
3456 789
= + - *
/
[ ]
( )
< >
{ }
Aa1@
012
001-FT-102
3456 789
012
0 1 2
001-FT-102
3456 789
= + - *
/
[ ]
( )
< >
{ }
Aa1@
012
001-FT-102
3456 789
012
0 1 2
001-FT-102
3456 789
= + - *
/
[ ]
( )
< >
{ }
Aa1@
012
001-FT-102
1496-1
1x
2x
4x
2x
1x
001-FT-101
1496-1
1.
2.
3.
4.
5.
6.
7.
8.
9.
Max.
Tag description
Ent. access code
Def. access code
/../Advanced setup
Tag description
/../Advanced setup
Ent. access code
Def. access code

8.3.9 Changing the parameters

For a description of the editing display - consisting of text editor and numeric editor ­with symbols →  41, for a description of the operating elements →  43
Example: Changing the tag name in the "Tag description" parameter from 001-FT-101 to 001-FT-102
Endress+Hauser 47
A message is displayed if the value entered is outside the permitted value range.
A0029563-EN
Page 48
Operation options Proline Promass F 200 HART
Ent.accesscode Invalidoroutofrangeinput value
Max:9999
Min:0
A0014049-EN

8.3.10 User roles and related access authorization

The two user roles "Operator" and "Maintenance" have different write access to the parameters if the customer defines a user-specific access code. This protects the device configuration via the local display from unauthorized access .
Access authorization to parameters: "Operator" user role
Access code status Read access Write access
An access code has not yet been defined (factory setting).
After an access code has been defined. ––
1)
1) Despite the defined access code, certain parameters can always be modified and thus are excepted from the write protection, as they do not affect the measurement. Refer to the "Write protection via access code" section
Access authorization to parameters: "Maintenance" user role
Access code status Read access Write access
An access code has not yet been defined (factory setting).
After an access code has been defined.
1) If an incorrect access code is entered, the user obtains the access rights of the "Operator" user role.
1)
The user role with which the user is currently logged on is indicated by the Access
status display parameter. Navigation path: Operation → Access status display

8.3.11 Disabling write protection via access code

If the -symbol appears on the local display in front of a parameter, the parameter is write-protected by a user-specific access code and its value cannot be changed at the moment using local operation →  91.
Parameter write protection via local operation can be disabled by entering the user-specific access code in the Enter access code parameter via the respective access option.
1. After you press , the input prompt for the access code appears.
2. Enter the access code.  The -symbol in front of the parameters disappears; all previously write-
protected parameters are now re-enabled.

8.3.12 Enabling and disabling the keypad lock

The keypad lock makes it possible to block access to the entire operating menu via local operation. As a result, it is no longer possible to navigate through the operating menu or change the values of individual parameters. Users can only read the measured values on the operational display.
48 Endress+Hauser
Page 49
Proline Promass F 200 HART Operation options
Local operation with mechanical push buttons (display module SD02)
Display module SD02: order characteristic "Display; Operation", option C
The keypad lock is switched on and off in the same way.
Switching on the keypad lock
The device is in the measured value display.
Press the  +  +  keys simultaneously.  The message Keylock on appears on the display: The keypad lock is switched on.
If the user attempts to access the operating menu while the keypad lock is active, the message Keylock on appears.
Switching off the keypad lock
The keypad lock is switched on.
Press the  +  +  keys simultaneously.  The message Keylock off appears on the display: The keypad lock is switched off.
Local operation with touch control (display module SD03)
Display module SD03: Order characteristic "Display; Operation", option E
The keypad lock is switched on and off via the context menu.
Switching on the keypad lock
The keypad lock is switched on automatically:
• Each time the device is restarted.
• If the device has not been operated for longer than one minute in the measured value display.
1. The device is in the measured value display. Press  for at least 2 seconds.  A context menu appears.
2. In the context menu, select the Keylock on option.  The keypad lock is switched on.
If the user attempts to access the operating menu while the keypad lock is active, the message Keylock on appears.
Switching off the keypad lock
1. The keypad lock is switched on. Press  for at least 2 seconds.  A context menu appears.
2. In the context menu, select the Keylock off option.  The keypad lock is switched off.

8.4 Access to the operating menu via the operating tool

The structure of the operating menu in the operating tools is the same as for operation via the local display.
Endress+Hauser 49
Page 50
Operation options Proline Promass F 200 HART
1 4 5 7
9
6 8
2
3
+
E
12
3

8.4.1 Connecting the operating tool

Via HART protocol
This communication interface is available in device versions with a HART output.
A0028746
 13 Options for remote operation via HART protocol (passive)
1 Control system (e.g. PLC) 2 Transmitter power supply unit, e.g. RN221N (with communication resistor) 3 Connection for Commubox FXA195 and Field Communicator 475 4 Field Communicator 475 5 Computer with Web browser (e.g. Internet Explorer) for accessing the integrated device Web server or
computer with operating tool (e.g. FieldCare, DeviceCare, AMS Device Manager, SIMATIC PDM) with COM
DTM "CDI Communication TCP/IP" 6 Commubox FXA195 (USB) 7 Field Xpert SFX350 or SFX370 8 VIATOR Bluetooth modem with connecting cable 9 Transmitter
Via service interface (CDI)
A0014019
1 Service interface (CDI = Endress+Hauser Common Data Interface) of the measuring device 2 Commubox FXA291 3 Computer with "FieldCare" operating tool with COM DTM "CDI Communication FXA291"
50 Endress+Hauser
Page 51
Proline Promass F 200 HART Operation options

8.4.2 Field Xpert SFX350, SFX370

Function scope
Field Xpert SFX350 and Field Xpert SFX370 are mobile computers for commissioning and maintenance. They enable efficient device configuration and diagnostics for HART and FOUNDATION fieldbus devices in the non-Ex area (SFX350, SFX370) and the Ex area (SFX370).
For details, see Operating Instructions BA01202S
Source for device description files
See data →  54

8.4.3 FieldCare

Function scope
FDT-based plant asset management tool from Endress+Hauser. It can configure all smart field devices in a system and helps you manage them. By using the status information, it is also a simple but effective way of checking their status and condition.
Access is via:
• HART protocol
• CDI service interface →  50 Typical functions:
• Configuring parameters of transmitters
• Loading and saving device data (upload/download)
• Documentation of the measuring point
• Visualization of the measured value memory (line recorder) and event logbook For additional information about FieldCare, see Operating Instructions BA00027S
and BA00059S
Source for device description files
See information →  54
Establishing a connection
For additional information, see Operating Instructions BA00027S and BA00059S
Endress+Hauser 51
Page 52
Operation options Proline Promass F 200 HART
6
5
32
1
Xxxxxx/…/…/
7
P
P
+
P
P
+
+
+
+
+
+
4
8 9
10
11
Xxxxxxx
GoodStatus:
Device tag:
Xxxxxxx
Device name:
Mass flow:
12.34
kg/h
Volume flow:
12.34
m /h³
Mass flow unit:
Volume flow unit:
kg/h
m /h³
Access status tooling
Operation
Setup
Xxxxxx
Mass flow unit Volume flow unit
Select medium
Device tag
Advanced setup
Diagnostics
Expert
Maintenance
kg/h m /h³
Xxxxxx
System units
User interface
1 Header 2 Picture of device 3 Device name 4 Tag name 5 Status area with status signal→  106 6 Display area for current measured values 7 Edit toolbar with additional functions such as save/restore, event list and create documentation 8 Navigation area with operating menu structure 9 Working area 10 Range of action 11 Status area

8.4.4 DeviceCare

Function scope
Tool to connect and configure Endress+Hauser field devices.
52 Endress+Hauser
The fastest way to configure Endress+Hauser field devices is with the dedicated "DeviceCare" tool. Together with the device type managers (DTMs) it presents a convenient, comprehensive solution.
For details, see Innovation Brochure IN01047S
Source for device description files
See information →  54

8.4.5 AMS Device Manager

Function scope
Program from Emerson Process Management for operating and configuring measuring devices via HART protocol.
A0021051-EN
Page 53
Proline Promass F 200 HART Operation options
Source for device description files
See data →  54

8.4.6 SIMATIC PDM

Function scope
SIMATIC PDM is a standardized, manufacturer-independent program from Siemens for the operation, configuration, maintenance and diagnosis of intelligent field devices via HART protocol.
Source for device description files
See data →  54

8.4.7 Field Communicator 475

Function scope
Industrial handheld terminal from Emerson Process Management for remote configuration and measured value display via HART protocol.
Source for device description files
See data →  54
Endress+Hauser 53
Page 54
System integration Proline Promass F 200 HART

9 System integration

9.1 Overview of device description files

9.1.1 Current version data for the device

Firmware version 01.04.zz • On the title page of the Operating instructions
• On the transmitter nameplate
• Firmware version Diagnostics → Device information → Firmware version
Release date of firmware version 06.2015 ---
Manufacturer ID 0x11 Manufacturer ID
Diagnostics → Device information → Manufacturer ID
Device type ID 0x54 Device type
Diagnostics → Device information → Device type
HART protocol revision 7 ---
Device revision 5 • On the transmitter nameplate
• Device revision Diagnostics → Device information → Device revision
For an overview of the different firmware versions for the device →  117

9.1.2 Operating tools

The suitable device description file for the individual operating tools is listed in the table below, along with information on where the file can be acquired.
Operating tool via HART protocol
FieldCare www.endress.com → Download Area
DeviceCare www.endress.com → Download Area
• Field Xpert SFX350
• Field Xpert SFX370
AMS Device Manager (Emerson Process Management)
SIMATIC PDM (Siemens)
Field Communicator 475 (Emerson Process Management)
Sources for obtaining device descriptions
• CD–ROM (contact Endress+Hauser)
• DVD (contact Endress+Hauser)
• CD–ROM (contact Endress+Hauser)
• DVD (contact Endress+Hauser)
Use update function of handheld terminal
www.endress.com → Download Area
www.endress.com → Download Area
Use update function of handheld terminal

9.2 Measured variables via HART protocol

The following measured variables (HART device variables) are assigned to the dynamic variables at the factory:
54 Endress+Hauser
Page 55
Proline Promass F 200 HART System integration
Dynamic variables Measured variables
(HART device variables)
Primary dynamic variable (PV) Mass flow
Secondary dynamic variable (SV) Totalizer 1
Tertiary dynamic variable (TV) Density
Quaternary dynamic variable (QV) Temperature
The assignment of the measured variables to the dynamic variables can be modified and assigned as desired via local operation and the operating tool using the following parameters:
• Expert → Communication → HART output → Output → Assign PV
• Expert → Communication → HART output → Output → Assign SV
• Expert → Communication → HART output → Output → Assign TV
• Expert → Communication → HART output → Output → Assign QV The following measured variables can be assigned to the dynamic variables:
Measured variables for PV (primary dynamic variable)
• Off
• Mass flow
• Volume flow
• Corrected volume flow
• Density
• Reference density
• Temperature
• Carrier pipe temperature
• Electronic temperature
• Oscillation frequency
• Oscillation amplitude
• Oscillation damping
• Signal asymmetry
Measured variables for SV, TV, QV (secondary, tertiary and quaternary dynamic variable)
• Mass flow
• Volume flow
• Corrected volume flow
• Density
• Reference density
• Temperature
• Carrier pipe temperature
• Electronic temperature
• Oscillation frequency
• Oscillation amplitude
• Oscillation damping
• Signal asymmetry
• External pressure
• Totalizer 1…3
Device variables
The device variables are permanently assigned. A maximum of 8 device variables can be transmitted:
• 0 = mass flow
• 1 = volume flow
• 2 = corrected volume flow
• 3 = density
• 4 = reference density
• 5 = temperature
• 6 = totalizer 1
Endress+Hauser 55
Page 56
System integration Proline Promass F 200 HART
• 7 = totalizer 2
• 8 = totalizer 3
• 9 = carrier pipe temperature
• 10 = electronic temperature
• 11 = oscillation damping
• 12 = oscillation frequency
• 13 = oscillation amplitude
• 14 = signal asymmetry
• 15 = external pressure

9.3 Other settings

Burst mode functionality in accordance with HART 7 Specification:
Navigation
"Expert" menu → Communication → HART output → Burst configuration → Burst configuration 1 to n
Burst configuration
Burst configuration 1 to n
Burst mode 1 to n
Burst command 1 to n
Burst variable 0
Burst variable 1
Burst variable 2
Burst variable 3
Burst variable 4
Burst variable 5
Burst variable 6
Burst variable 7
→  57
→  57
→  57
→  57
→  57
→  57
→  57
→  57
→  57
→  57
Burst trigger mode
Burst trigger level
Min. update period
Max. update period
→  57
→  57
→  58
→  58
56 Endress+Hauser
Page 57
Proline Promass F 200 HART System integration
Parameter overview with brief description
Parameter Description Selection / User entry Factory setting
Burst mode 1 to n Activate the HART burst mode for burst
message X.
Burst command 1 to n Select the HART command that is sent to the
HART master.
Burst variable 0 For HART command 9 and 33: select the
HART device variable or the process variable.
Burst variable 1 For HART command 9 and 33: select the
HART device variable or the process variable.
Burst variable 2 For HART command 9 and 33: select the
HART device variable or the process variable.
Burst variable 3 For HART command 9 and 33: select the
HART device variable or the process variable.
Burst variable 4 For HART command 9 and 33: select the
HART device variable or the process variable.
Burst variable 5 For HART command 9 and 33: select the
HART device variable or the process variable.
Burst variable 6 For HART command 9 and 33: select the
HART device variable or the process variable.
Burst variable 7 For HART command 9 and 33: select the
HART device variable or the process variable.
Burst trigger mode Select the event that triggers burst messageX.• Continuous
Burst trigger level Enter the burst trigger value.
Together with the option selected in the Burst trigger mode parameter the burst trigger value determines the time of burst message X.
• Off
• On
• Command 1
• Command 2
• Command 3
• Command 9
• Command 33
• Command 48
• Mass flow
• Volume flow
• Corrected volume flow
• Density
• Reference density
• Temperature
• Carrier pipe temperature
• Electronic temperature
• Oscillation damping
• Oscillation frequency
• Oscillation amplitude 0
• Signal asymmetry
• Totalizer 1
• Totalizer 2
• Totalizer 3
• External pressure
• Percent Of Range
• Measured current
• Primary variable (PV)
• Secondary variable (SV)
• Tertiary variable (TV)
• Quaternary variable (QV)
• Not used
See the Burst variable 0 parameter.
See the Burst variable 0 parameter.
See the Burst variable 0 parameter.
See the Burst variable 0 parameter.
See the Burst variable 0 parameter.
See the Burst variable 0 parameter.
See the Burst variable 0 parameter.
• Window
• Rising
• Falling
• On change
Positive floating-point number –
Off
Command 2
Mass flow
Not used
Not used
Not used
Not used
Not used
Not used
Not used
Continuous
Endress+Hauser 57
Page 58
System integration Proline Promass F 200 HART
Parameter Description Selection / User entry Factory setting
Min. update period Enter the minimum time span between two
burst commands of burst message X.
Max. update period Enter the maximum time span between two
burst commands of burst message X.
Positive integer 1 000 ms
Positive integer 2 000 ms
58 Endress+Hauser
Page 59
Proline Promass F 200 HART Commissioning
X X X X X X XX X
20.50
Operation Setup
Main menu
0104-1
Display language
English
Español Français
Display language
English
Deutsch
Ã
0104-1
Ã
Español Français
Display language
English
Deutsch
0104-1
Betrieb Setup
Hauptmenü
Sprache
Deutsch
0104-1
XXXX
1.
2.
3.
4.

10 Commissioning

10.1 Function check

Before commissioning the measuring device:
Make sure that the post-installation and post-connection checks have been performed.
• "Post-installation check" checklist →  26
• "Post-connection check" checklist →  34

10.2 Switching on the measuring device

After a successful function check, switch on the measuring device.
 After a successful startup, the local display switches automatically from the startup
display to the operational display.
If nothing appears on the local display or a diagnostic message is displayed, refer to the section on "Diagnostics and troubleshooting" →  101.

10.3 Setting the operating language

Factory setting: English or ordered local language
Endress+Hauser 59
 14 Taking the example of the local display

10.4 Configuring the measuring device

• The Setup menuwith its guided wizards contains all the parameters needed for standard operation.
• Navigation to the Setup menu
A0029420
Page 60
Commissioning Proline Promass F 200 HART
X X X X X X XX X
20.50
0104-1
XXXXXXXXX XXXXXXXXX
mA
1.
2.
3.
Display/operat.
Setup
Diagnostic
Display/operat. Setup
Main menu
English
Main menu
Medium selection
/ ../Setup
Display language
A0032222-EN
 15 Taking the example of the local display
Setup
Device tag
Select medium
System units
Current output 1 to n
Pulse/frequency/switch output
Display
Output conditioning
Low flow cut off
Partially filled pipe detection
Advanced setup
→  61
→  62
→  65
→  66
→  73
→  75
→  78
→  79
→  80

10.4.1 Defining the tag name

60 Endress+Hauser
To enable fast identification of the measuring point within the system, you can enter a unique designation using the Device tag parameter and thus change the factory setting.
Page 61
Proline Promass F 200 HART Commissioning
1
XXXXXXXXX
A0029422
 16 Header of the operational display with tag name
1 Tag name
Enter the tag name in the "FieldCare" operating tool →  52
Navigation
"Setup" menu → Device tag
Parameter overview with brief description
Parameter Description User entry Factory setting
Device tag Enter the name for the measuring point. Max. 32 characters, such as
letters, numbers or special characters (e.g. @, %, /).
Promass
Endress+Hauser 61
Page 62
Commissioning Proline Promass F 200 HART

10.4.2 Selecting and setting the medium

The Medium selection wizard systematically guides the user through all the parameters that must be configured in order to select and set the medium.
Navigation
"Setup" menu → Medium selection
Select medium
Select medium
Select gas type
Reference sound velocity
Temperature coefficient sound velocity
Pressure compensation
Pressure value
Parameter overview with brief description
Parameter Prerequisite Description Selection / User
Select medium Select medium type. • Liquid
• Gas
Select gas type In the Select medium
parameter the Gas option is selected.
Reference sound velocity In the Select gas type
parameter the Others option is selected.
Temperature coefficient sound velocity
Pressure compensation Select pressure compensation
Pressure value In the Pressure compensation
In the Select gas type parameter the Others option is selected.
parameter the Fixed value option is selected.
Select measured gas type. Gas type choose list Air
Enter sound velocity of gas at 0 °C (32 °F).
Enter temperature coefficient for the gas sound velocity.
type.
Enter process pressure to be used for pressure correction.
1 to 99 999.9999 m/s
Positive floating­point number
• Off
• Fixed value
• External value
Positive floating­point number
→  62
→  62
→  62
→  62
→  62
→  62
Factory setting
entry
Liquid
331.5 m/s
0.61 (m/s)/K
Off
Country-specific:
• 1.01 bar a
• 14.7 psi a

10.4.3 Setting the system units

In the System units submenu the units of all the measured values can be set.
Depending on the device version, not all submenus and parameters are available in every device. The selection can vary depending on the order code.
62 Endress+Hauser
Page 63
Proline Promass F 200 HART Commissioning
Navigation
"Setup" menu → System units
System units
Mass flow unit
Mass unit
Volume flow unit
Volume unit
Corrected volume flow unit
Corrected volume unit
Density unit
Reference density unit
Temperature unit
Length unit
Pressure unit
→  63
→  63
→  63
→  63
→  64
→  64
→  64
→  64
→  64
→  64
→  64
Parameter overview with brief description
Parameter Description Selection Factory setting
Mass flow unit Select mass flow unit.
Result
The selected unit applies for:
• Output
• Low flow cut off
• Simulation process variable
Mass unit Select mass unit. Unit choose list Country-specific:
Volume flow unit Select volume flow unit.
Result
The selected unit applies for:
• Output
• Low flow cut off
• Simulation process variable
Volume unit Select volume unit. Unit choose list Country-specific:
Unit choose list Country-specific:
• kg/h
• lb/min
• kg
• lb
Unit choose list Country-specific:
• l/h
• gal/min (us)
• l (DN > 150 (6"): m³)
• gal (us)
Endress+Hauser 63
Page 64
Commissioning Proline Promass F 200 HART
Parameter Description Selection Factory setting
Corrected volume flow unit Select corrected volume flow unit.
Result
The selected unit applies for: Corrected volume flow parameter (→  95)
Corrected volume unit Select corrected volume unit. Unit choose list Country-specific:
Density unit Select density unit.
Result
The selected unit applies for:
• Output
• Simulation process variable
• Density adjustment (Expert menu)
Reference density unit Select reference density unit. Unit choose list Country-dependent
Temperature unit Select temperature unit.
Result
The selected unit applies for:
• Minimum value
• Maximum value
• Maximum value
• Minimum value
• Average value
• Minimum value
• Maximum value
• Minimum value
• Maximum value
• Reference temperature
Length unit Select length unit for nominal diameter. Unit choose list Country-specific:
Pressure unit Select process pressure unit.
Result
The unit is taken from:
Pressure value parameter (→  62)
External pressure parameter
Unit choose list Country-specific:
• Nl/h
• Sft³/min
• Nl
• Sft³
Unit choose list Country-specific:
• kg/l
• lb/ft³
• kg/Nl
• lb/Sft³
Unit choose list Country-specific:
• °C
• °F
• mm
• in
Unit choose list Country-specific:
• bar a
• psi a
64 Endress+Hauser
Page 65
Proline Promass F 200 HART Commissioning

10.4.4 Configuring the current output

The Current output wizard guides you systematically through all the parameters that have to be set for configuring the current output.
Navigation
"Setup" menu → Current output 1 to n
Current output 1 to n
Assign current output
Current span
4 mA value
20 mA value
Fixed current
Failure mode
Failure current
Parameter overview with brief description
Parameter Prerequisite Description Selection / User
Assign current output Select process variable for
current output.
Corrected volume flow calculation Select reference density for
calculating the corrected volume flow.
Fixed reference density The Fixed reference density
option is selected in the
Corrected volume flow calculation parameter
parameter.
Enter fixed value for reference density.
entry
• Off
• Mass flow
• Volume flow
• Corrected volume flow
• Density
• Reference density
• Temperature
• Carrier pipe temperature
• Electronic temperature
• Oscillation frequency
• Oscillation amplitude
• Oscillation damping
• Signal asymmetry
• Fixed reference density
• Calculated reference density
Positive floating­point number
→  65
→  66
→  66
→  66
→  66
→  66
Factory setting
Mass flow
Calculated reference density
1 kg/Nl
Endress+Hauser 65
Page 66
Commissioning Proline Promass F 200 HART
Parameter Prerequisite Description Selection / User
Linear expansion coefficient The Calculated reference
density option is selected in the Corrected volume flow calculation parameter
parameter.
Square expansion coefficient The Calculated reference
density option is selected in the Corrected volume flow calculation parameter parameter.
Reference temperature The Calculated reference
density option is selected in the Corrected volume flow calculation parameter parameter.
Current span Select current range for
4 mA value One of the following options is
selected in the Current span parameter (→  66):
• 4...20 mA NAMUR
• 4...20 mA US
• 4...20 mA
20 mA value One of the following options is
selected in the Current span parameter (→  66):
• 4...20 mA NAMUR
• 4...20 mA US
• 4...20 mA
Failure mode One of the following options is
selected in the Assign current output parameter (→  65):
• Mass flow
• Volume flow
• Corrected volume flow
• Density
• Reference density
• Temperature
• Carrier pipe temperature
• Electronic temperature
• Oscillation frequency
• Oscillation amplitude
• Oscillation damping
• Signal asymmetry
One of the following options is selected in the Current span parameter (→  66):
• 4...20 mA NAMUR
• 4...20 mA US
• 4...20 mA
Failure current In the Failure mode
parameter, the Defined value option is selected.
Enter linear, medium-specific expansion coefficient for calculating the reference density.
For media with a non-linear expansion pattern: enter the quadratic, medium-specific expansion coefficient for calculating the reference density.
Enter reference temperature for calculating the reference density.
process value output and upper/lower level for alarm signal.
Enter 4 mA value. Signed floating-point
Enter 20 mA value. Signed floating-point
Define output behavior in alarm condition.
Enter current output value in alarm condition.
Factory setting
entry
Signed floating-point number
Signed floating-point number
–273.15 to 99 999 °C Country-specific:
• 4...20 mA NAMUR
• 4...20 mA US
• 4...20 mA
• Fixed current
number
number
• Min.
• Max.
• Last valid value
• Actual value
• Defined value
3.59 to 22.5 mA 22.5 mA
0.0
0.0
• +20 °C
• +68 °F
Country-specific:
• 4...20 mA NAMUR
• 4...20 mA US
Country-specific:
• 0 kg/h
• 0 lb/min
Depends on country and nominal diameter
Max.

10.4.5 Configuring the pulse/frequency/switch output

The Pulse/frequency/switch output wizard guides you systematically through all the parameters that can be set for configuring the selected output type.
66 Endress+Hauser
Page 67
Proline Promass F 200 HART Commissioning
Navigation
"Setup" menu → Pulse/frequency/switch output
Pulse/frequency/switch output
Operating mode
Parameter overview with brief description
Parameter Description Selection Factory setting
Operating mode Define the output as a pulse, frequency or
switch output.
Configuring the pulse output
Navigation
"Setup" menu → Pulse/frequency/switch output
Pulse/frequency/switch output
Operating mode
Assign pulse output
• Pulse
• Frequency
• Switch
→  67
Pulse
→  67
→  67
Value per pulse
Pulse width
Failure mode
Invert output signal
Parameter overview with brief description
Parameter Prerequisite Description Selection / User
Operating mode Define the output as a pulse,
frequency or switch output.
Assign pulse output In the Operating mode
parameter, the Pulse option is selected.
Select process variable for pulse output.
entry
• Pulse
• Frequency
• Switch
• Off
• Mass flow
• Volume flow
• Corrected volume flow
→  68
→  68
→  68
→  68
Factory setting
Pulse
Off
Endress+Hauser 67
Page 68
Commissioning Proline Promass F 200 HART
Parameter Prerequisite Description Selection / User
Value per pulse In the Operating mode
parameter, the Pulse option is selected and one of the following options is selected in the Assign pulse output parameter (→  67):
• Mass flow
• Volume flow
• Corrected volume flow
Pulse width In the Operating mode
parameter, the Pulse option is selected and one of the following options is selected in the Assign pulse output parameter (→  67):
• Mass flow
• Volume flow
• Corrected volume flow
Failure mode In the Operating mode
parameter, the Pulse option is selected and one of the following options is selected in the Assign pulse output parameter (→  67):
• Mass flow
• Volume flow
• Corrected volume flow
Invert output signal Invert the output signal. • No
Enter measured value at which a pulse is output.
Define time width of the output pulse.
Define output behavior in alarm condition.
Signed floating-point number
5 to 2 000 ms 100 ms
• Actual value
• No pulses
• Yes
Factory setting
entry
Depends on country and nominal diameter
No pulses
No
Configuring the frequency output
Navigation
"Setup" menu → Pulse/frequency/switch output
Pulse/frequency/switch output
Operating mode
Assign frequency output
Minimum frequency value
Maximum frequency value
Measuring value at minimum frequency
Measuring value at maximum frequency
→  69
→  69
→  69
→  70
→  70
→  70
Failure mode
→  71
68 Endress+Hauser
Page 69
Proline Promass F 200 HART Commissioning
Failure frequency
Invert output signal
Parameter overview with brief description
Parameter Prerequisite Description Selection / User
Operating mode Define the output as a pulse,
frequency or switch output.
Assign frequency output In the Operating mode
parameter (→  67), the Frequency option is selected.
Minimum frequency value In the Operating mode
parameter, the Frequency option is selected and one of the following options is selected in the Assign frequency output parameter (→  69):
• Mass flow
• Volume flow
• Corrected volume flow
• Density
• Reference density
• Temperature
• Carrier pipe temperature
• Electronic temperature
• Oscillation frequency
• Oscillation amplitude
• Oscillation damping
• Signal asymmetry
Select process variable for frequency output.
Enter minimum frequency. 0 to 1 000 Hz 0 Hz
entry
• Pulse
• Frequency
• Switch
• Off
• Mass flow
• Volume flow
• Corrected volume flow
• Density
• Reference density
• Temperature
• Carrier pipe temperature
• Electronic temperature
• Oscillation frequency
• Oscillation amplitude
• Oscillation damping
• Signal asymmetry
→  71
→  71
Factory setting
Pulse
Off
Endress+Hauser 69
Page 70
Commissioning Proline Promass F 200 HART
Parameter Prerequisite Description Selection / User
Maximum frequency value In the Operating mode
parameter, the Frequency option is selected and one of the following options is selected in the Assign frequency output parameter (→  69):
• Mass flow
• Volume flow
• Corrected volume flow
• Density
• Reference density
• Temperature
• Carrier pipe temperature
• Electronic temperature
• Oscillation frequency
• Oscillation amplitude
• Oscillation damping
• Signal asymmetry
Measuring value at minimum frequency
Measuring value at maximum frequency
In the Operating mode parameter, the Frequency option is selected and one of the following options is selected in the Assign frequency output parameter (→  69):
• Mass flow
• Volume flow
• Corrected volume flow
• Density
• Reference density
• Temperature
• Carrier pipe temperature
• Electronic temperature
• Oscillation frequency
• Oscillation amplitude
• Oscillation damping
• Signal asymmetry
In the Operating mode parameter, the Frequency option is selected and one of the following options is selected in the Assign frequency output parameter (→  69):
• Mass flow
• Volume flow
• Corrected volume flow
• Density
• Reference density
• Temperature
• Carrier pipe temperature
• Electronic temperature
• Oscillation frequency
• Oscillation amplitude
• Oscillation damping
• Signal asymmetry
Factory setting
entry
Enter maximum frequency. 0 to 1 000 Hz 1 000 Hz
Enter measured value for minmum frequency.
Enter measured value for maximum frequency.
Signed floating-point number
Signed floating-point number
Depends on country and nominal diameter
Depends on country and nominal diameter
70 Endress+Hauser
Page 71
Proline Promass F 200 HART Commissioning
Parameter Prerequisite Description Selection / User
Failure mode In the Operating mode
parameter (→  67), the Frequency option is selected and one of the following options is selected in the
Assign frequency output
parameter (→  69):
• Mass flow
• Volume flow
• Corrected volume flow
• Density
• Reference density
• Temperature
• Carrier pipe temperature
• Electronic temperature
• Oscillation frequency
• Oscillation amplitude
• Oscillation damping
• Signal asymmetry
Failure frequency In the Operating mode
parameter (→  67), the Frequency option is selected and one of the following options is selected in the
Assign frequency output
parameter (→  69):
• Mass flow
• Volume flow
• Corrected volume flow
• Density
• Reference density
• Temperature
• Carrier pipe temperature
• Electronic temperature
• Oscillation frequency
• Oscillation amplitude
• Oscillation damping
• Signal asymmetry
Invert output signal Invert the output signal. • No
Define output behavior in alarm condition.
Enter frequency output value in alarm condition.
• Actual value
• Defined value
• 0 Hz
0.0 to 1 250.0 Hz 0.0 Hz
• Yes
Factory setting
entry
0 Hz
No
Configuring the switch output
Navigation
"Setup" menu → Pulse/frequency/switch output
Pulse/frequency/switch output
Operating mode
Switch output function
Assign diagnostic behavior
Assign limit
Assign flow direction check
→  72
→  72
→  72
→  72
→  72
Endress+Hauser 71
Page 72
Commissioning Proline Promass F 200 HART
Assign status
Switch-on value
Switch-off value
Switch-on delay
Switch-off delay
Failure mode
Invert output signal
Parameter overview with brief description
Parameter Prerequisite Description Selection / User
Operating mode Define the output as a pulse,
frequency or switch output.
Switch output function In the Operating mode
parameter the Switch option is selected.
Assign diagnostic behavior • In the Operating mode
parameter, the Switch option is selected.
• In the Switch output
function parameter, the Diagnostic behavior option
is selected.
Assign limit • In the Operating mode
parameter, the Switch option is selected.
• In the Switch output
function parameter, the Limit option is selected.
Assign flow direction check • The Switch option is
selected in the Operating mode parameter.
• The Flow direction check option is selected in the
Switch output function
parameter.
Select function for switch output.
Select diagnostic behavior for switch output.
Select process variable for limit function.
Select process variable for flow direction monitoring.
entry
• Pulse
• Frequency
• Switch
• Off
• On
• Diagnostic behavior
• Limit
• Flow direction check
• Status
• Alarm
• Alarm or warning
• Warning
• Mass flow
• Volume flow
• Corrected volume flow
• Density
• Reference density
• Temperature
• Totalizer 1
• Totalizer 2
• Totalizer 3
• Mass flow
• Volume flow
• Corrected volume flow
→  73
→  73
→  73
→  73
→  73
→  73
→  73
Factory setting
Pulse
Off
Alarm
Mass flow
Mass flow
72 Endress+Hauser
Page 73
Proline Promass F 200 HART Commissioning
Parameter Prerequisite Description Selection / User
Assign status • The Switch option is
selected in the Operating mode parameter.
• The Status option is selected in the Switch output function parameter.
Switch-on value • In the Operating mode
parameter, the Switch option is selected.
• In the Switch output
function parameter, the Limit option is selected.
Switch-off value • In the Operating mode
parameter, the Switch option is selected.
• In the Switch output
function parameter, the Limit option is selected.
Switch-on delay • The Switch option is
selected in the Operating mode parameter.
• The Limit option is selected in the Switch output function parameter.
Switch-off delay • The Switch option is
selected in the Operating mode parameter.
• The Limit option is selected in the Switch output function parameter.
Failure mode Define output behavior in
Invert output signal Invert the output signal. • No
Select device status for switch output.
Enter measured value for the switch-on point.
Enter measured value for the switch-off point.
Define delay for the switch-on of status output.
Define delay for the switch-off of status output.
alarm condition.
• Partially filled pipe detection
• Low flow cut off
Signed floating-point number
Signed floating-point number
0.0 to 100.0 s 0.0 s
0.0 to 100.0 s 0.0 s
• Actual status
• Open
• Closed
• Yes
Factory setting
entry
Partially filled pipe detection
Country-specific:
• 0 kg/h
• 0 lb/min
Country-specific:
• 0 kg/h
• 0 lb/min
Open
No

10.4.6 Configuring the local display

The Display wizard guides you systematically through all the parameters that can configured for configuring the local display.
Navigation
"Setup" menu → Display
Display
Format display
Value 1 display
0% bargraph value 1
100% bargraph value 1
Value 2 display
Endress+Hauser 73
→  74
→  74
→  74
→  74
→  74
Page 74
Commissioning Proline Promass F 200 HART
Value 3 display
0% bargraph value 3
100% bargraph value 3
Value 4 display
Parameter overview with brief description
Parameter Prerequisite Description Selection / User
Format display A local display is provided. Select how measured values
are shown on the display.
Value 1 display A local display is provided. Select the measured value that
is shown on the local display.
0% bargraph value 1 A local display is provided. Enter 0% value for bar graph
display.
100% bargraph value 1 A local display is provided. Enter 100% value for bar
graph display.
Value 2 display A local display is provided. Select the measured value that
is shown on the local display.
Value 3 display A local display is provided. Select the measured value that
is shown on the local display.
0% bargraph value 3 A selection was made in the
Value 3 display parameter.
100% bargraph value 3 A selection was made in the
Value 3 display parameter.
Value 4 display A local display is provided. Select the measured value that
Enter 0% value for bar graph display.
Enter 100% value for bar graph display.
is shown on the local display.
entry
• 1 value, max. size
• 1 bargraph + 1 value
• 2 values
• 1 value large + 2 values
• 4 values
• None
• Mass flow
• Volume flow
• Corrected volume flow
• Density
• Reference density
• Temperature
• Totalizer 1
• Totalizer 2
• Totalizer 3
• Current output 1
• Current output 2
Signed floating-point number
Signed floating-point number
For the picklist, see the Value 1 display parameter
For the picklist, see the Value 1 display parameter (→  74)
Signed floating-point number
Signed floating-point number
For the picklist, see the Value 1 display parameter (→  74)
→  74
→  74
→  74
→  74
Factory setting
1 value, max. size
Mass flow
*
Country-specific:
• 0 kg/h
• 0 lb/min
Depends on country and nominal diameter
None
None
Country-specific:
• 0 kg/h
• 0 lb/min
0
None
* Visibility depends on order options or device settings
74 Endress+Hauser
Page 75
Proline Promass F 200 HART Commissioning

10.4.7 Configuring the output conditioning

The Output conditioning wizard guides you systematically through all the parameters that have to be set for configuring the output conditioning.
Navigation
"Setup" menu → Output conditioning
Output conditioning
Level of flow fluctuation
Display damping
Damping output 1
Damping output 2
Damping output 2
Measuring mode output 1
Measuring mode output 2
Measuring mode output 2
Measuring mode output 2
Operating mode totalizer 1
Operating mode totalizer 2
→  76
→  76
→  76
→  76
→  76
→  76
→  76
→  76
→  76
→  76
→  76
Operating mode totalizer 3
Assign process variable
On value low flow cutoff
Off value low flow cutoff
Pressure shock suppression
→  76
→  76
→  77
→  77
→  77
Endress+Hauser 75
Page 76
Commissioning Proline Promass F 200 HART
Parameter overview with brief description
Parameter Prerequisite Description Selection / User
Level of flow fluctuation Select fluctuation level of
measured value.
Display damping Set display reaction time to
fluctuations in the measured value.
Damping output 1 Set the reaction time of the
output signal of the current output to fluctuations in the measured value.
Damping output 2 The measuring device has a
second current output.
Damping output 2 The measuring device has a
pulse/frequency/switch output.
Measuring mode output 1 Select measuring mode for
Measuring mode output 2 Select measuring mode for
Measuring mode output 2 Select measuring mode for
Measuring mode output 2 Select measuring mode for
Operating mode totalizer 1 Select totalizer calculation
Operating mode totalizer 2 Select totalizer calculation
Operating mode totalizer 3 Select totalizer calculation
Assign process variable Select process variable for low
Set the reaction time of the output signal of the second current output to fluctuations in the measured value.
Set the reaction time of the output signal of the frequency output to fluctuations in the measured value.
output.
output.
output.
output.
mode.
mode.
mode.
flow cut off.
entry
• Slight
• Moderate
• Strong
0.0 to 999.9 s 0.0 s
0 to 999.9 s 1 s
0 to 999.9 s 1 s
0 to 999.9 s 1 s
• Forward flow
• Forward/Reverse flow
• Reverse flow compensation
• Forward flow
• Forward/Reverse flow
• Reverse flow compensation
• Forward flow
• Forward/Reverse flow
• Reverse flow
• Reverse flow compensation
• Forward flow
• Forward/Reverse flow
• Reverse flow
• Reverse flow compensation
• Net flow total
• Forward flow total
• Reverse flow total
• Net flow total
• Forward flow total
• Reverse flow total
• Net flow total
• Forward flow total
• Reverse flow total
• Off
• Mass flow
• Volume flow
• Corrected volume flow
Moderate
Forward flow
Forward flow
Forward flow
Forward flow
Net flow total
Net flow total
Net flow total
Mass flow
Factory setting
76 Endress+Hauser
Page 77
Proline Promass F 200 HART Commissioning
Parameter Prerequisite Description Selection / User
On value low flow cutoff One of the following options is
selected in the Assign process variable parameter (→  76):
• Mass flow
• Volume flow
• Corrected volume flow
Off value low flow cutoff One of the following options is
selected in the Assign process variable parameter (→  76):
• Mass flow
• Volume flow
• Corrected volume flow
Pressure shock suppression One of the following options is
selected in the Assign process variable parameter (→  76):
• Mass flow
• Volume flow
• Corrected volume flow
Enter on value for low flow cut off.
Enter off value for low flow cut off.
Enter time frame for signal suppression (= active pressure shock suppression).
entry
Positive floating­point number
0 to 100.0 % 50 %
0 to 100 s 0 s
Depends on country and nominal diameter
Factory setting
Endress+Hauser 77
Page 78
Commissioning Proline Promass F 200 HART

10.4.8 Configuring the low flow cut off

The Low flow cut off wizard systematically guides the user through all the parameters that must be set to configure low flow cut off.
Navigation
"Setup" menu → Low flow cut off
Low flow cut off
Assign process variable
On value low flow cutoff
Off value low flow cutoff
Pressure shock suppression
Parameter overview with brief description
Parameter Prerequisite Description Selection / User
Assign process variable Select process variable for low
flow cut off.
On value low flow cutoff One of the following options is
selected in the Assign process variable parameter (→  76):
• Mass flow
• Volume flow
• Corrected volume flow
Off value low flow cutoff One of the following options is
selected in the Assign process variable parameter (→  76):
• Mass flow
• Volume flow
• Corrected volume flow
Pressure shock suppression One of the following options is
selected in the Assign process variable parameter (→  76):
• Mass flow
• Volume flow
• Corrected volume flow
Enter on value for low flow cut off.
Enter off value for low flow cut off.
Enter time frame for signal suppression (= active pressure shock suppression).
entry
• Off
• Mass flow
• Volume flow
• Corrected volume flow
Positive floating­point number
0 to 100.0 % 50 %
0 to 100 s 0 s
Mass flow
Depends on country and nominal diameter
→  78
→  78
→  78
→  78
Factory setting
78 Endress+Hauser
Page 79
Proline Promass F 200 HART Commissioning

10.4.9 Configuring the partial filled pipe detection

The Partial filled pipe detection wizard guides you systematically through all parameters that have to be set for configuring the monitoring of the pipe filling.
Navigation
"Setup" menu → Partially filled pipe detection
Partially filled pipe detection
Assign process variable
Low value partial filled pipe detection
High value partial filled pipe detection
Response time part. filled pipe detect.
Parameter overview with brief description
Parameter Prerequisite Description Selection / User
Assign process variable Select process variable for
partially filled pipe detection.
Low value partial filled pipe detection
High value partial filled pipe detection
Response time part. filled pipe detect.
One of the following options is selected in the Assign process variable parameter:
• Density
• Reference density
One of the following options is selected in the Assign process variable parameter:
• Density
• Reference density
One of the following options is selected in the Assign process variable parameter:
• Density
• Reference density
Enter lower limit value for deactivating partialy filled pipe detection.
Enter upper limit value for deactivating partialy filled pipe detection.
Enter time before diagnostic message is displayed for partially filled pipe detection.
entry
• Off
• Density
• Reference density
Positive floating­point number
Signed floating-point number
0 to 100 s 1 s
Off
200
6 000
→  79
→  79
→  79
→  79
Factory setting
Endress+Hauser 79
Page 80
Commissioning Proline Promass F 200 HART
X X X X X X XX X
20.50
0104-1
0092-1
****
XXXXXXXXX
XXXXXXXXX
XXXXXXXXX XXXXXXXXX
mA
1.
2.
3.
4.
5.
Display/operat.
Setup
Diagnostic
Display/operat.
Setup
Main menu
English
Main menu
Medium selection
/ ../Setup
/ ../Setup
Advanced setup
Ent. access code
/ ../Advanced setup
Def. access code
Device tag
Display language

10.5 Advanced settings

The Advanced setup submenu together with its submenus contains parameters for specific settings.
Navigation to the "Advanced setup" submenu
Navigation
"Setup" menu → Advanced setup
Advanced setup
Enter access code
Sensor adjustment
Totalizer 1 to n
SIL confirmation
Deactivate SIL
Display
A0032223-EN
→  81
→  82
→  84
80 Endress+Hauser
Page 81
Proline Promass F 200 HART Commissioning
Heartbeat setup
Configuration backup display
Administration
→  87
→  86

10.5.1 Carrying out a sensor adjustment

The Sensor adjustment submenu contains parameters that pertain to the functionality of the sensor.
Navigation
"Setup" menu → Advanced setup → Sensor adjustment
Sensor adjustment
Installation direction
Zero point adjustment
Parameter overview with brief description
→  81
→  81
Parameter Description Selection Factory setting
Installation direction Set sign of flow direction to match the
direction of the arrow on the sensor.
Zero point adjustment
All measuring devices are calibrated in accordance with state-of-the-art technology. Calibration takes place under reference conditions →  133. Therefore, a zero point adjustment in the field is generally not required.
Experience shows that zero point adjustment is advisable only in special cases:
• To achieve maximum measuring accuracy even with low flow rates
• Under extreme process or operating conditions (e.g. very high process temperatures or very high-viscosity fluids).
Navigation
"Setup" menu → Advanced setup → Sensor adjustment → Zero point adjustment
Zero point adjustment
Zero point adjustment control
Adjustment in progress
• Flow in arrow direction
• Flow against arrow direction
Flow in arrow direction
→  82
→  82
Endress+Hauser 81
Page 82
Commissioning Proline Promass F 200 HART
Parameter overview with brief description
Parameter Prerequisite Description Selection / User
interface
Zero point adjustment control Start zero point adjustment. • Cancel
• Busy
• Zero point adjust failure
• Start
Adjustment in progress The Start option is selected in
the Zero point adjustment control parameter.
0 to 100 %

10.5.2 Configuring the totalizer

In the "Totalizer 1 to n" submenu the individual totalizer can be configured.
Navigation
"Setup" menu → Advanced setup → Totalizer 1 to n
Totalizer 1 to n
Assign process variable
Unit totalizer
Factory setting
Cancel
→  82
→  82
Totalizer operation mode
Failure mode
Parameter overview with brief description
Parameter Prerequisite Description Selection Factory setting
Assign process variable Select process variable for
totalizer.
Unit totalizer One of the following options is
selected in the Assign process variable parameter (→  82)Totalizer 1 to n submenu:
• Volume flow
• Mass flow
• Corrected volume flow
Select process variable totalizer unit.
→  83
→  83
• Off
• Volume flow
• Mass flow
• Corrected volume flow
Unit choose list Country-specific:
Mass flow
• l
• gal (us)
82 Endress+Hauser
Page 83
Proline Promass F 200 HART Commissioning
Parameter Prerequisite Description Selection Factory setting
Totalizer operation mode One of the following options is
selected in the Assign process variable parameter (→  82)Totalizer 1 to n submenu:
• Volume flow
• Mass flow
• Corrected volume flow
Failure mode One of the following options is
selected in the Assign process variable parameter (→  82)Totalizer 1 to n submenu:
• Volume flow
• Mass flow
• Corrected volume flow
Select totalizer calculation mode.
Define totalizer behavior in alarm condition.
• Net flow total
• Forward flow total
• Reverse flow total
• Stop
• Actual value
• Last valid value
Net flow total
Stop
Endress+Hauser 83
Page 84
Commissioning Proline Promass F 200 HART

10.5.3 Carrying out additional display configurations

In the Display submenu you can set all the parameters associated with the configuration of the local display.
Navigation
"Setup" menu → Advanced setup → Display
Display
Format display
Value 1 display
0% bargraph value 1
100% bargraph value 1
Decimal places 1
Value 2 display
Decimal places 2
Value 3 display
0% bargraph value 3
100% bargraph value 3
Decimal places 3
→  85
→  85
→  85
→  85
→  85
→  85
→  85
→  85
→  85
→  85
→  85
Value 4 display
Decimal places 4
Language
Display interval
Display damping
Header
Header text
Separator
Backlight
→  85
→  86
→  86
→  86
→  86
→  86
→  86
→  86
→  86
84 Endress+Hauser
Page 85
Proline Promass F 200 HART Commissioning
Parameter overview with brief description
Parameter Prerequisite Description Selection / User
Format display A local display is provided. Select how measured values
are shown on the display.
Value 1 display A local display is provided. Select the measured value that
is shown on the local display.
0% bargraph value 1 A local display is provided. Enter 0% value for bar graph
display.
100% bargraph value 1 A local display is provided. Enter 100% value for bar
graph display.
Decimal places 1 A measured value is specified
in the Value 1 display parameter.
Value 2 display A local display is provided. Select the measured value that
Decimal places 2 A measured value is specified
in the Value 2 display parameter.
Value 3 display A local display is provided. Select the measured value that
0% bargraph value 3 A selection was made in the
Value 3 display parameter.
100% bargraph value 3 A selection was made in the
Value 3 display parameter.
Decimal places 3 A measured value is specified
in the Value 3 display parameter.
Value 4 display A local display is provided. Select the measured value that
Select the number of decimal places for the display value.
is shown on the local display.
Select the number of decimal places for the display value.
is shown on the local display.
Enter 0% value for bar graph display.
Enter 100% value for bar graph display.
Select the number of decimal places for the display value.
is shown on the local display.
entry
• 1 value, max. size
• 1 bargraph + 1 value
• 2 values
• 1 value large + 2 values
• 4 values
• None
• Mass flow
• Volume flow
• Corrected volume flow
• Density
• Reference density
• Temperature
• Totalizer 1
• Totalizer 2
• Totalizer 3
• Current output 1
• Current output 2
Signed floating-point number
Signed floating-point number
• x
• x.x
• x.xx
• x.xxx
• x.xxxx
For the picklist, see the Value 1 display parameter
• x
• x.x
• x.xx
• x.xxx
• x.xxxx
For the picklist, see the Value 1 display parameter (→  74)
Signed floating-point number
Signed floating-point number
• x
• x.x
• x.xx
• x.xxx
• x.xxxx
For the picklist, see the Value 1 display parameter (→  74)
Factory setting
1 value, max. size
Mass flow
*
Country-specific:
• 0 kg/h
• 0 lb/min
Depends on country and nominal diameter
x.xx
None
x.xx
None
Country-specific:
• 0 kg/h
• 0 lb/min
0
x.xx
None
Endress+Hauser 85
Page 86
Commissioning Proline Promass F 200 HART
Parameter Prerequisite Description Selection / User
Decimal places 4 A measured value is specified
in the Value 4 display parameter.
Select the number of decimal places for the display value.
• x
• x.x
• x.xx
• x.xxx
• x.xxxx
Language A local display is provided. Set display language. • English
• Deutsch
• Français
• Español
• Italiano
• Nederlands
• Portuguesa
• Polski
• русский язык (Russian)
• Svenska
• Türkçe
中文 (Chinese)
日本語 (Japanese)
한국어 (Korean)
• Bahasa Indonesia
• tiếng Việt (Vietnamese)
• čeština (Czech)
Display interval A local display is provided. Set time measured values are
1 to 10 s 5 s shown on display if display alternates between values.
Display damping A local display is provided. Set display reaction time to
0.0 to 999.9 s 0.0 s fluctuations in the measured value.
Header A local display is provided. Select header contents on local
display.
Header text In the Header parameter, the
Enter display header text. Max. 12 characters
Free text option is selected.
• Device tag
• Free text
such as letters, numbers or special characters (e.g. @, %, /)
Separator A local display is provided. Select decimal separator for
displaying numerical values.
Backlight Order code for "Display;
operation", option E "SD03 4-
Switch the local display backlight on and off.
• . (point)
• , (comma)
• Disable
• Enable
line, illum.; touch control + data backup function"
entry
*
*
*
*
*
*
*
Factory setting
x.xx
English (alternatively, the ordered language is preset in the device)
*
*
*
*
*
*
*
*
*
Device tag
------------
. (point)
Disable
* Visibility depends on order options or device settings

10.5.4 Using parameters for device administration

The Administration submenu systematically guides the user through all the parameters that can be used for device administration purposes.
86 Endress+Hauser
Page 87
Proline Promass F 200 HART Commissioning
Navigation
"Setup" menu → Advanced setup → Administration
Administration
Define access code
Define access code
Confirm access code
Device reset
Parameter overview with brief description
Parameter Description User entry / Selection Factory setting
Define access code Restrict write-access to parameters to
protect the configuration of the device against unintentional changes via the local display.
Confirm access code Confirm the entered access code. 0 to 9 999 0
Device reset Reset the device configuration - either
entirely or in part - to a defined state.
0 to 9 999 0
• Cancel
• To factory defaults
• To delivery settings
• Restart device

10.6 Configuration management

→  87
→  87
→  87
Cancel
After commissioning, you can save the current device configuration, copy it to another measuring point or restore the previous device configuration.
You can do so using the Configuration management parameter and the related options found in the Configuration backup display submenu.
Navigation
"Setup" menu → Advanced setup → Configuration backup display
Configuration backup display
Operating time
Last backup
Configuration management
Comparison result
→  88
→  88
→  88
→  88
Endress+Hauser 87
Page 88
Commissioning Proline Promass F 200 HART
Parameter overview with brief description
Parameter Prerequisite Description User interface / 
Operating time Indicates how long the device
has been in operation.
Last backup A local display is provided. Indicates when the last data
backup was saved to the display module.
Configuration management A local display is provided. Select action for managing the
device data in the display module.
Comparison result A local display is provided. Comparison between present
device data and display backup.
Selection
Days (d), hours (h), minutes (m) and seconds (s)
Days (d), hours (h), minutes (m) and seconds (s)
• Cancel
• Execute backup
• Restore
• Duplicate
• Compare
• Clear backup data
• Settings identical
• Settings not
identical
• No backup
available
• Backup settings
corrupt
• Check not done
• Dataset
incompatible
Factory setting
Cancel
Check not done

10.6.1 Function scope of the "Configuration management" parameter

Options Description
Cancel No action is executed and the user exits the parameter.
Execute backup A backup copy of the current device configuration is saved from the integrated
HistoROM to the display module of the device. The backup copy includes the transmitter data of the device.
Restore The last backup copy of the device configuration is restored from the display
module to the device's integrated HistoROM. The backup copy includes the transmitter data of the device.
Compare The device configuration saved in the display module is compared with the current
device configuration of the integrated HistoROM.
Duplicate The transmitter configuration from another device is duplicated to the device using
the display module.
Clear backup data The backup copy of the device configuration is deleted from the display module of
the device.
Integrated HistoROM
A HistoROM is a "non-volatile" device memory in the form of an EEPROM. While this action is in progress, the configuration cannot be edited via the local
display and a message on the processing status appears on the display.

10.7 Simulation

The Simulation submenu enables you to simulate, without a real flow situation, various process variables in the process and the device alarm mode and to verify downstream signal chains (switching valves or closed-control loops).
88 Endress+Hauser
Page 89
Proline Promass F 200 HART Commissioning
Navigation
"Diagnostics" menu → Simulation
Simulation
Assign simulation process variable
Process variable value
Simulation current output 1 to n
Value current output 1 to n
Frequency output simulation
Frequency value
Pulse output simulation
Pulse value
Switch output simulation
Switch status
Simulation device alarm
→  89
→  89
→  90
→  90
→  90
→  90
→  90
→  90
→  90
→  90
→  90
Diagnostic event category
Diagnostic event simulation
Parameter overview with brief description
Parameter Prerequisite Description Selection / User
Assign simulation process variable Select a process variable for
the simulation process that is activated.
Process variable value One of the following options is
selected in the Assign
simulation process variable
parameter (→  89):
• Mass flow
• Volume flow
• Corrected volume flow
• Density
• Reference density
• Temperature
Enter the simulation value for the selected process variable.
entry
• Off
• Mass flow
• Volume flow
• Corrected volume flow
• Density
• Reference density
• Temperature
Depends on the process variable selected
→  90
→  90
Factory setting
Off
0
Endress+Hauser 89
Page 90
Commissioning Proline Promass F 200 HART
Parameter Prerequisite Description Selection / User
Simulation current output 1 to n Switch the simulation of the
current output on and off.
Value current output 1 to n In the Simulation current
output 1 to n parameter, the On option is selected.
Frequency output simulation In the Operating mode
parameter, the Frequency option is selected.
Frequency value In the Frequency output
simulation parameter, the On option is selected.
Pulse output simulation In the Operating mode
parameter, the Pulse option is selected.
Pulse value In the Pulse output
simulation parameter (→  90), the Down- counting value option is selected.
Switch output simulation In the Operating mode
parameter, the Switch option is selected.
Switch status In the Switch output
simulation parameter (→  90) Switch output
simulation 1 to n parameter Switch output simulation 1 to n parameter, the On option is
selected.
Simulation device alarm Switch the device alarm on and
Diagnostic event category Select a diagnostic event
Diagnostic event simulation Select a diagnostic event to
Enter the current value for simulation.
Switch the simulation of the frequency output on and off.
Enter the frequency value for the simulation.
Set and switch off the pulse output simulation.
For Fixed value option:
Pulse width parameter (→  68) defines the pulse width of the pulses output.
Enter the number of pulses for simulation.
Switch the simulation of the switch output on and off.
Select the status of the status output for the simulation.
off.
category.
simulate this event.
Factory setting
entry
• Off
• On
3.59 to 22.5 mA 3.59 mA
• Off
• On
0.0 to 1 250.0 Hz 0.0 Hz
• Off
• Fixed value
• Down-counting value
0 to 65 535 0
• Off
• On
• Open
• Closed
• Off
• On
• Sensor
• Electronics
• Configuration
• Process
• Off
• Diagnostic event picklist (depends on the category selected)
Off
Off
Off
Off
Open
Off
Process
Off

10.8 Protecting settings from unauthorized access

The following options exist for protecting the configuration of the measuring device from unintentional modification after commissioning:
• Write protection via access code
• Write protection via write protection switch
• Write protection via keypad lock →  48
90 Endress+Hauser
Page 91
Proline Promass F 200 HART Commissioning

10.8.1 Write protection via access code

The effects of the user-specific access code are as follows:
• Via local operation, the parameters for the measuring device configuration are write­protected and their values can no longer be changed.
• Device access is protected via the Web browser, as are the parameters for the measuring device configuration.
Defining the access code via local display
1. Navigate to the Enter access code parameter.
2. Define a max. 16-digit character string comprising numbers, letters and special characters as the access code.
3. Enter the access code again in the to confirm the code.  The -symbol appears in front of all write-protected parameters.
The device automatically locks the write-protected parameters again if a key is not pressed for 10 minutes in the navigation and editing view. The device locks the write-protected parameters automatically after 60 s if the user skips back to the operational display mode from the navigation and editing view.
• If parameter write protection is activated via an access code, it can also only be deactivated via this access code →  48.
• The user role with which the user is currently logged on via the local display is indicated by the →  48 Access status display parameter. Navigation path: Operation → Access status display
Parameters which can always be modified via the local display
Certain parameters that do not affect the measurement are excepted from parameter write protection via the local display. Despite the user-specific access code, they can always be modified, even if the other parameters are locked.
Parameters for configuring
the local display
Language Format display Control Totalizer
Contrast display Preset value
Display interval Reset all totalizers
Parameters for configuring
the totalizer

10.8.2 Write protection via write protection switch

Unlike parameter write protection via a user-specific access code, this allows write access to the entire operating menu - except for the "Contrast display" parameter - to be locked.
The parameter values are now read only and cannot be edited any more (exception "Contrast display" parameter):
• Via local display
• Via service interface (CDI)
• Via HART protocol
Endress+Hauser 91
Page 92
Commissioning Proline Promass F 200 HART
+
E
3 mm
XXXXXXXXXXXXX
WPCDI
1.2. 3.
4.
ON
OFF
A0032230
1. Loosen the securing clamp.
2. Unscrew the electronics compartment cover.
3. Pull out the display module with a gentle rotational movement. To make it easier to access the write protection switch, attach the display module to the edge of the electronics compartment.
 Display module is attached to the edge of the electronics compartment.
A0032236
92 Endress+Hauser
Page 93
Proline Promass F 200 HART Commissioning
X X X X X X XX X
20.50
XX
XX
4. Setting the write protection switch (WP) on the main electronics module to the ON position enables hardware write protection. Setting the write protection switch (WP) on the main electronics module to the OFF position (factory setting) disables hardware write protection.
 If the hardware write protection is enabled: The Hardware locked option is
displayed in the Locking status parameter . In addition, on the local display the -symbol appears in front of the parameters in the header of the operational display and in the navigation view.
A0029425
If the hardware write protection is disabled: No option is displayed in the Locking status parameter . On the local display, the -symbol disappears from in front of the parameters in the header of the operational display and in the navigation view.
5. Feed the cable into the gap between the housing and main electronics module and plug the display module into the electronics compartment in the desired direction until it engages.
6. Reverse the removal procedure to reassemble the transmitter.
Endress+Hauser 93
Page 94
Operation Proline Promass F 200 HART

11 Operation

11.1 Reading the device locking status

Device active write protection: Locking status parameter
Operation → Locking status
Function scope of the "Locking status" parameter
Options Description
None The access status displayed in the Access status display parameter applies
→  48. Only appears on local display.
Hardware locked The DIP switch for hardware locking is activated on the main electronics module.
This locks write access to the parameters (e.g. via local display or operating tool).
SIL locked The SIL mode is enabled. This locks write access to the parameters (e.g. via local
display or operating tool).
Temporarily locked Write access to the parameters is temporarily locked on account of internal
processes running in the device (e.g. data upload/download, reset etc.). Once the internal processing has been completed, the parameters can be changed once again.

11.2 Adjusting the operating language

Detailed information:
• To configure the operating language →  59
• For information on the operating languages supported by the measuring device →  143

11.3 Configuring the display

Detailed information:
• On the basic settings for the local display →  73
• On the advanced settings for the local display →  84

11.4 Reading measured values

With the Measured values submenu, it is possible to read all the measured values.

11.4.1 Process variables

The Process variables submenu contains all the parameters needed to display the current measured values for each process variable.
Navigation
"Diagnostics" menu → Measured values → Process variables
Process variables
Mass flow
Volume flow
→  95
→  95
94 Endress+Hauser
Page 95
Proline Promass F 200 HART Operation
Corrected volume flow
Density
Reference density
Temperature
Parameter overview with brief description
Parameter Description User interface
Mass flow Displays the mass flow currently measured.
Dependency
The unit is taken from the Mass flow unit parameter
Volume flow Displays the volume flow currently measured.
Dependency
The unit is taken from the Volume flow unit parameter
Corrected volume flow Displays the corrected volume flow currently calculated.
Dependency
The unit is taken from the Corrected volume flow unit parameter
Density Displays the density or specific density currently measured.
Dependency
The unit is taken from the Density unit parameter
Reference density Displays the density at the reference temperature.
Dependency
The unit is taken from the Reference density unit parameter
Temperature Displays the temperature currently measured.
Dependency
The unit is taken from the Temperature unit parameter
→  95
→  95
→  95
→  95
Signed floating-point number
Signed floating-point number
Signed floating-point number
Positive floating-point number
Positive floating-point number
Positive floating-point number

11.4.2 "Totalizer" submenu

The Totalizer submenu contains all the parameters needed to display the current measured values for every totalizer.
Navigation
"Diagnostics" menu → Measured values → Totalizer
Totalizer
Totalizer value 1 to n
Totalizer overflow 1 to n
Endress+Hauser 95
→  96
→  96
Page 96
Operation Proline Promass F 200 HART
Parameter overview with brief description
Parameter Prerequisite Description User interface
Totalizer value 1 to n One of the following options is selected
in the Assign process variable parameter (→  82)Totalizer 1 to n submenu:
• Volume flow
• Mass flow
• Corrected volume flow
Totalizer overflow 1 to n One of the following options is selected
in the Assign process variable parameter (→  82)Totalizer 1 to n submenu:
• Volume flow
• Mass flow
• Corrected volume flow
Displays the current totalizer counter value.
Displays the current totalizer overflow. Integer with sign
Signed floating-point number

11.4.3 Output values

The Output values submenu contains all the parameters needed to display the current measured values for every output.
Navigation
"Diagnostics" menu → Measured values → Output values
Output values
Output current 1
Measured current 1
Terminal voltage 1
Output current 2
Pulse output
Output frequency
Switch status
Parameter overview with brief description
→  96
→  96
→  97
→  97
→  97
Parameter Prerequisite Description User interface
Output current 1 Displays the current value currently
calculated for the current output.
Measured current 1 Displays the current value currently
measured for the current output.
Terminal voltage 1 Displays the current terminal voltage
that is applied at the output.
Output current 2 Displays the current value currently
calculated for the current output.
3.59 to 22.5 mA
0 to 30 mA
0.0 to 50.0 V
3.59 to 22.5 mA
96 Endress+Hauser
Page 97
Proline Promass F 200 HART Operation
Parameter Prerequisite Description User interface
Pulse output In the Operating mode parameter, the
Pulse option is selected.
Output frequency In the Operating mode parameter, the
Frequency option is selected.
Switch status The Switch option is selected in the
Operating mode parameter.
Displays the pulse frequency currently output.
Displays the value currently measured for the frequency output.
Displays the current switch output status.
Positive floating-point number
0 to 1 250 Hz
• Open
• Closed

11.5 Adapting the measuring device to the process conditions

The following are available for this purpose:
• Basic settings using theSetup menu (→  59)
• Advanced settings using theAdvanced setup submenu (→  80)

11.6 Performing a totalizer reset

The totalizers are reset in the Operation submenu:
• Control Totalizer
• Reset all totalizers
Navigation
"Operation" menu → Totalizer handling
Totalizer handling
Control Totalizer 1 to n
Preset value 1 to n
Reset all totalizers
Parameter overview with brief description
Parameter Prerequisite Description Selection / User
Control Totalizer 1 to n One of the following options is
selected in the Assign process variable parameter (→  82)Totalizer 1 to n submenu:
• Volume flow
• Mass flow
• Corrected volume flow
Preset value 1 to n One of the following options is
selected in the Assign process variable parameter (→  82)Totalizer 1 to n submenu:
• Volume flow
• Mass flow
• Corrected volume flow
Reset all totalizers Reset all totalizers to 0 and
Control totalizer value. • Totalize
Specify start value for totalizer.
Dependency
The unit of the selected
process variable is specified for the totalizer in the Unit totalizer parameter (→  82).
start.
entry
• Reset + hold
• Preset + hold
• Reset + totalize
• Preset + totalize
• Hold
Signed floating-point number
• Cancel
• Reset + totalize
→  97
→  97
→  97
Factory setting
Totalize
Country-specific:
• 0 l
• 0 gal (us)
Cancel
Endress+Hauser 97
Page 98
Operation Proline Promass F 200 HART

11.6.1 Function scope of the "Control Totalizer" parameter

Options Description
Totalize The totalizer is started or continues running.
Reset + hold The totaling process is stopped and the totalizer is reset to 0.
Preset + hold The totaling process is stopped and the totalizer is set to its defined start value
from the Preset value parameter.
Reset + totalize The totalizer is reset to 0 and the totaling process is restarted.
Preset + totalize The totalizer is set to the defined start value from the Preset value parameter and
the totaling process is restarted.
Hold Totalizing is stopped.

11.6.2 Function scope of the "Reset all totalizers" parameter

Options Description
Cancel No action is executed and the user exits the parameter.
Reset + totalize Resets all totalizers to 0 and restarts the totaling process. This deletes all the flow
values previously totalized.

11.7 Showing data logging

The Extended HistoROM application package must be enabled in the device (order option) for the Data logging submenu to appear. This contains all the parameters for the measured value history.
Data logging is also available via:
• Plant Asset Management Tool FieldCare →  51.
• Web browser
Function range
• A total of 1000 measured values can be stored
• 4 logging channels
• Adjustable logging interval for data logging
• Display of the measured value trend for each logging channel in the form of a chart
A0016357
 17 Chart of a measured value trend
• x-axis: depending on the number of channels selected displays 250 to 1000 measured values of a process variable.
• y-axis: displays the approximate measured value span and constantly adapts this to the ongoing measurement.
If the length of the logging interval or the assignment of the process variables to the channels is changed, the content of the data logging is deleted.
98 Endress+Hauser
Page 99
Proline Promass F 200 HART Operation
Navigation
"Diagnostics" menu → Data logging
Data logging
Assign channel 1...4
Logging interval
Clear logging data
Data logging
Logging delay
Data logging control
Data logging status
Entire logging duration
Parameter overview with brief description
Parameter Prerequisite Description Selection / User
Assign channel 1 to n The Extended HistoROM
application package is available.
Logging interval The Extended HistoROM
application package is available.
Clear logging data The Extended HistoROM
application package is available.
Data logging Select the data logging
Assign process variable to logging channel.
Define the logging interval for data logging. This value defines the time interval between the individual data points in the memory.
Clear the entire logging data. • Cancel
method.
Factory setting
entry / User
interface
• Off
• Mass flow
• Volume flow
• Corrected volume flow
• Density
• Reference density
• Temperature
• Carrier pipe temperature
• Electronic temperature
• Oscillation frequency
• Oscillation amplitude
• Oscillation damping
• Signal asymmetry
• Current output 1
1.0 to 3 600.0 s 10.0 s
• Clear data
• Overwriting
• Not overwriting
Off
Cancel
Overwriting
→  99
→  99
→  99
→  99
→  100
→  100
→  100
→  100
Endress+Hauser 99
Page 100
Operation Proline Promass F 200 HART
Parameter Prerequisite Description Selection / User
Logging delay In the Data logging
parameter, the Not overwriting option is selected.
Data logging control In the Data logging
parameter, the Not overwriting option is selected.
Data logging status In the Data logging
parameter, the Not overwriting option is selected.
Entire logging duration In the Data logging
parameter, the Not overwriting option is selected.
Enter the time delay for measured value logging.
Start and stop measured value logging.
Displays the measured value logging status.
Displays the total logging duration.
entry / User
interface
0 to 999 h 0 h
• None
• Delete + start
• Stop
• Done
• Delay active
• Active
• Stopped
Positive floating­point number
None
Done
0 s
Factory setting
100 Endress+Hauser