Endress Hauser F 200 Operating Instructions

BA01686D/06/EN/03.20 71483612 2020-06-01
Valid as of version
Products Solutions Services
Operating Instructions
Proline Prowirl F 200 HART
Vortex flowmeter
Proline Prowirl F 200 HART
• Make sure the document is stored in a safe place such that it is always available when working on or with the device.
• To avoid danger to individuals or the facility, read the "Basic safety instructions" section carefully, as well as all other safety instructions in the document that are specific to working procedures.
• The manufacturer reserves the right to modify technical data without prior notice. Your Endress+Hauser Sales Center will supply you with current information and updates to these instructions.
2 Endress+Hauser
Proline Prowirl F 200 HART Table of contents

Table of contents

1 About this document ................ 6
1.1 Document function ..................... 6
1.2 Symbols .............................. 6
1.2.1 Safety symbols .................. 6
1.2.2 Electrical symbols ................ 6
1.2.3 Communication symbols ........... 6
1.2.4 Tool symbols .................... 7
1.2.5 Symbols for
certain types of information ......... 7
1.2.6 Symbols in graphics ............... 7
1.3 Documentation ........................ 8
1.3.1 Standard documentation ........... 8
1.3.2 Supplementary device-dependent
documentation .................. 8
1.4 Registered trademarks ................... 8
2 Safety instructions .................. 9
2.1 Requirements for the personnel ............ 9
2.2 Designated use ........................ 9
2.3 Workplace safety ...................... 10
2.4 Operational safety ..................... 10
2.5 Product safety ........................ 10
2.6 IT security ........................... 11
2.7 Device-specific IT security ................ 11
2.7.1 Protecting access via hardware write
protection ..................... 11
2.7.2 Protecting access via a password .... 11
2.7.3 Access via fieldbus ............... 11
3 Product description ................ 12
3.1 Product design ........................ 12
4 Incoming acceptance and product
identification ..................... 14
4.1 Incoming acceptance ................... 14
4.2 Product identification ................... 14
4.2.1 Transmitter nameplate ........... 15
4.2.2 Sensor nameplate ............... 16
4.2.3 Pressure measuring cells nameplate .. 19
4.2.4 Symbols on measuring device ...... 19
5 Storage and transport ............. 20
5.1 Storage conditions ..................... 20
5.2 Transporting the product ................ 20
5.2.1 Measuring devices without lifting
lugs ......................... 20
5.2.2 Measuring devices with lifting lugs .. 21
5.2.3 Transporting with a fork lift ........ 21
5.3 Packaging disposal ..................... 21
6 Installation ....................... 22
6.1 Installation conditions .................. 22
6.1.1 Mounting position ............... 22
6.1.2 Environment and process
requirements .................. 27
6.1.3 Special mounting instructions ...... 28
6.2 Mounting the measuring device ........... 29
6.2.1 Required tools .................. 29
6.2.2 Preparing the measuring device ..... 29
6.2.3 Mounting the sensor ............. 29
6.2.4 Mounting the pressure measuring
unit ......................... 30
6.2.5 Mounting the transmitter of the
remote version ................. 32
6.2.6 Turning the transmitter housing .... 33
6.2.7 Turning the display module ........ 34
6.3 Post-installation check .................. 34
7 Electrical connection .............. 36
7.1 Connection conditions .................. 36
7.1.1 Required tools .................. 36
7.1.2 Connecting cable requirements ..... 36
7.1.3 Connecting cable for remote
version ....................... 37
7.1.4 Terminal assignment ............. 38
7.1.5 Requirements for the supply unit .... 40
7.1.6 Preparing the measuring device ..... 41
7.2 Connecting the measuring device .......... 41
7.2.1 Connecting the compact version ..... 42
7.2.2 Connecting the remote version ..... 43
7.2.3 Connecting the connecting cable for
the pressure measuring cell ........ 48
7.2.4 Ensuring potential equalization ..... 48
7.3 Ensuring the degree of protection .......... 48
7.4 Post-connection check .................. 48
8 Operation options ................. 50
8.1 Overview of operation options ............ 50
8.2 Structure and function of the operating
menu .............................. 51
8.2.1 Structure of the operating menu .... 51
8.2.2 Operating philosophy ............ 52
8.3 Access to the operating menu via the local
display ............................. 53
8.3.1 Operational display .............. 53
8.3.2 Navigation view ................ 54
8.3.3 Editing view ................... 56
8.3.4 Operating elements .............. 58
8.3.5 Opening the context menu ......... 58
8.3.6 Navigating and selecting from list ... 60
8.3.7 Calling the parameter directly ...... 60
8.3.8 Calling up help text .............. 61
8.3.9 Changing the parameters ......... 62
Endress+Hauser 3
Table of contents Proline Prowirl F 200 HART
8.3.10 User roles and related access
authorization .................. 63
8.3.11 Disabling write protection via access
code ......................... 63
8.3.12 Enabling and disabling the keypad
lock ......................... 64
8.4 Access to the operating menu via the
operating tool ........................ 64
8.4.1 Connecting the operating tool ...... 64
8.4.2 Field Xpert SFX350, SFX370 ....... 65
8.4.3 FieldCare ..................... 66
8.4.4 DeviceCare .................... 67
8.4.5 AMS Device Manager ............ 67
8.4.6 SIMATIC PDM .................. 68
8.4.7 Field Communicator 475 .......... 68
9 System integration ................ 69
9.1 Overview of device description files ......... 69
9.1.1 Current version data for the device ... 69
9.1.2 Operating tools ................. 69
9.2 Measured variables via HART protocol ...... 69
9.3 Other settings ........................ 71
10 Commissioning .................... 74
10.1 Function check ....................... 74
10.2 Switching on the measuring device ......... 74
10.3 Setting the operating language ............ 74
10.4 Configuring the measuring device .......... 75
10.4.1 Defining the tag name ............ 75
10.4.2 Setting the system units .......... 76
10.4.3 Selecting and setting the medium ... 80
10.4.4 Configuring the current input ...... 83
10.4.5 Configuring the current output ..... 85
10.4.6 Configuring the pulse/frequency/
switch output .................. 86
10.4.7 Configuring the local display ....... 91
10.4.8 Configuring the output
conditioning ................... 93
10.4.9 Configuring the low flow cut off ..... 93
10.5 Advanced settings ..................... 95
10.5.1 Setting the medium properties ...... 96
10.5.2 Performing external compensation . 109
10.5.3 Carrying out a sensor adjustment ... 111
10.5.4 Configuring the totalizer ......... 114
10.5.5 Carrying out additional display
configurations ................. 116
10.5.6 Configuration management ....... 119
10.5.7 Using parameters for device
administration ................ 120
10.6 Configuration management ............. 120
10.6.1 Function scope of the "Configuration
management" parameter ......... 121
10.7 Simulation .......................... 122
10.8 Protecting settings from unauthorized
access ............................. 124
10.8.1 Write protection via access code ... 124
10.8.2 Write protection via write protection
switch ....................... 125
10.9 Application-specific commissioning ....... 126
10.9.1 Steam application .............. 126
10.9.2 Liquid application .............. 127
10.9.3 Gas applications ............... 128
10.9.4 Calculation of the measured
variables ..................... 131
11 Operation ....................... 136
11.1 Reading the device locking status ......... 136
11.2 Adjusting the operating language ......... 136
11.3 Configuring the display ................ 136
11.4 Reading measured values ............... 136
11.4.1 Process variables ............... 137
11.4.2 "Totalizer" submenu ............. 140
11.4.3 Input values .................. 140
11.4.4 Output values ................. 141
11.5 Adapting the measuring device to the process
conditions .......................... 142
11.6 Performing a totalizer reset ............. 142
11.6.1 Function scope of the "Control
Totalizer" parameter ............ 142
11.6.2 Function scope of the "Reset all
totalizers" parameter ............ 143
11.7 Showing data logging ................. 143
12 Diagnostics and troubleshooting .. 146
12.1 General troubleshooting ................ 146
12.2 Diagnostic information on local display ..... 148
12.2.1 Diagnostic message ............. 148
12.2.2 Calling up remedial measures ..... 150
12.3 Diagnostic information in FieldCare or
DeviceCare ......................... 150
12.3.1 Diagnostic options .............. 150
12.3.2 Calling up remedy information .... 152
12.4 Adapting the diagnostic information ...... 152
12.4.1 Adapting the diagnostic behavior ... 152
12.4.2 Adapting the status signal ........ 153
12.5 Overview of diagnostic information ....... 153
12.5.1 Operating conditions for displaying
the following diagnostics
information .................. 157
12.5.2 Emergency mode in the event of
pressure compensation .......... 157
12.5.3 Emergency mode in event of
temperature compensation ....... 157
12.6 Pending diagnostic events .............. 158
12.7 Diagnostic list ....................... 158
12.8 Event logbook ....................... 159
12.8.1 Reading out the event logbook ..... 159
12.8.2 Filtering the event logbook ....... 159
12.8.3 Overview of information events .... 160
12.9 Resetting the measuring device .......... 161
12.9.1 Function scope of the "Device reset"
parameter .................... 161
12.10 Device information ................... 161
12.11 Firmware history ..................... 163
4 Endress+Hauser
Proline Prowirl F 200 HART Table of contents
13 Maintenance .................... 164
13.1 Maintenance tasks .................... 164
13.1.1 Exterior cleaning ............... 164
13.1.2 Interior cleaning ............... 164
13.1.3 Replacing seals ................ 164
13.1.4 Adjusting the pressure measuring
cell ......................... 164
13.2 Measuring and test equipment ........... 165
13.3 Endress+Hauser services ............... 165
14 Repair ........................... 166
14.1 General notes ....................... 166
14.1.1 Repair and conversion concept ..... 166
14.1.2 Notes for repair and conversion .... 166
14.2 Spare parts ......................... 166
14.3 Endress+Hauser services ............... 167
14.4 Return ............................. 167
14.5 Disposal ........................... 167
14.5.1 Removing the measuring device .... 167
14.5.2 Disposing of the measuring device .. 168
15 Accessories ...................... 169
15.1 Device-specific accessories .............. 169
15.1.1 For the transmitter ............. 169
15.1.2 For the sensor ................. 170
15.2 Communication-specific accessories ....... 170
15.3 Service-specific accessories .............. 171
15.4 System components ................... 172
16 Technical data ................... 173
16.1 Application ......................... 173
16.2 Function and system design ............. 173
16.3 Input .............................. 173
16.4 Output ............................ 180
16.5 Power supply ........................ 183
16.6 Performance characteristics ............. 186
16.7 Installation ......................... 190
16.8 Environment ........................ 191
16.9 Process ............................ 192
16.10 Mechanical construction ............... 194
16.11 Operability ......................... 203
16.12 Certificates and approvals .............. 205
16.13 Application packages .................. 206
16.14 Accessories ......................... 206
16.15 Supplementary documentation ........... 207
Index ................................. 209
Endress+Hauser 5
About this document Proline Prowirl F 200 HART
DANGER
WARNING
CAUTION
NOTICE

1 About this document

1.1 Document function

These Operating Instructions contain all the information that is required in various phases of the life cycle of the device: from product identification, incoming acceptance and storage, to mounting, connection, operation and commissioning through to troubleshooting, maintenance and disposal.

1.2 Symbols

1.2.1 Safety symbols

This symbol alerts you to a dangerous situation. Failure to avoid this situation will result in serious or fatal injury.
This symbol alerts you to a dangerous situation. Failure to avoid this situation can result in serious or fatal injury.
This symbol alerts you to a dangerous situation. Failure to avoid this situation can result in minor or medium injury.
This symbol contains information on procedures and other facts which do not result in personal injury.

1.2.2 Electrical symbols

Symbol Meaning
Direct current
Alternating current
Direct current and alternating current
Ground connection
A grounded terminal which, as far as the operator is concerned, is grounded via a grounding system.
Protective Earth (PE)
A terminal which must be connected to ground prior to establishing any other connections.
The ground terminals are situated inside and outside the device:
• Inner ground terminal: Connects the protectiv earth to the mains supply.
• Outer ground terminal: Connects the device to the plant grounding system.

1.2.3 Communication symbols

Symbol Meaning
Wireless Local Area Network (WLAN)
Communication via a wireless, local network.
6 Endress+Hauser
Proline Prowirl F 200 HART About this document
A
1.
1.
-
.

1.2.4 Tool symbols

Symbol Meaning
Flat blade screwdriver
Allen key
Open-ended wrench

1.2.5 Symbols for certain types of information

Symbol Meaning
Permitted
Procedures, processes or actions that are permitted.
Preferred
Procedures, processes or actions that are preferred.
Forbidden
Procedures, processes or actions that are forbidden.
Tip
Indicates additional information.
Reference to documentation.
Reference to page.
Reference to graphic.
Notice or individual step to be observed.
, 2., 3.… Series of steps.
Result of a step.
Help in the event of a problem.
Visual inspection.

1.2.6 Symbols in graphics

Symbol Meaning
1, 2, 3, ... Item numbers
, 2., 3., … Series of steps
A, B, C, ... Views
A-A, B-B, C-C, ... Sections
Hazardous area
Safe area (non-hazardous area)
Flow direction
Endress+Hauser 7
About this document Proline Prowirl F 200 HART

1.3 Documentation

For an overview of the scope of the associated Technical Documentation, refer to the following:
W@M Device Viewer (www.endress.com/deviceviewer): Enter the serial number from nameplate
Endress+Hauser Operations App: Enter the serial number from the nameplate or scan the 2D matrix code (QR code) on the nameplate
Detailed list of the individual documents along with the documentation code →  207

1.3.1 Standard documentation

Document type Purpose and content of the document
Technical Information Planning aid for your device
The document contains all the technical data on the device and provides an overview of the accessories and other products that can be ordered for the device.
Sensor Brief Operating Instructions Guides you quickly to the 1st measured value - Part 1
The Sensor Brief Operating Instructions are aimed at specialists with responsibility for installing the measuring device.
• Incoming acceptance and product identification
• Storage and transport
• Installation
Transmitter Brief Operating Instructions
Description of Device Parameters Reference for your parameters
Guides you quickly to the 1st measured value - Part 2
The Transmitter Brief Operating Instructions are aimed at specialists with responsibility for commissioning, configuring and parameterizing the measuring device (until the first measured value).
• Product description
• Installation
• Electrical connection
• Operation options
• System integration
• Commissioning
• Diagnostic information
The document provides a detailed explanation of each individual parameter in the Expert operating menu. The description is aimed at those who work with the device over the entire life cycle and perform specific configurations.

1.3.2 Supplementary device-dependent documentation

Additional documents are supplied depending on the device version ordered: Always comply strictly with the instructions in the supplementary documentation. The supplementary documentation is an integral part of the device documentation.

1.4 Registered trademarks

HART®
Registered trademark of the FieldComm Group, Austin, Texas, USA
KALREZ®, VITON®
Registered trademarks of DuPont Performance Elastomers L.L.C., Wilmington, DE USA
GYLON®
Registered trademark of Garlock Sealing Technologies, Palmyar, NY, USA
8 Endress+Hauser
Proline Prowirl F 200 HART Safety instructions

2 Safety instructions

2.1 Requirements for the personnel

The personnel for installation, commissioning, diagnostics and maintenance must fulfill the following requirements:
Trained, qualified specialists must have a relevant qualification for this specific function
and task. Are authorized by the plant owner/operator.
Are familiar with federal/national regulations.
Before starting work, read and understand the instructions in the manual and
supplementary documentation as well as the certificates (depending on the application). Follow instructions and comply with basic conditions.
The operating personnel must fulfill the following requirements:
Are instructed and authorized according to the requirements of the task by the facility's
owner-operator. Follow the instructions in this manual.

2.2 Designated use

Application and media
The measuring device described in this manual is intended only for flow measurement of liquids with a minimum conductivity of 20 µS/cm.
Depending on the version ordered, the measuring device can also measure potentially explosive, flammable, poisonous and oxidizing media.
Measuring devices for use in hazardous areas, in hygienic applications or where there is an increased risk due to process pressure, are labeled accordingly on the nameplate.
To ensure that the measuring device remains in proper condition for the operation time:
Keep within the specified pressure and temperature range.
Only use the measuring device in full compliance with the data on the nameplate and
the general conditions listed in the Operating Instructions and supplementary documentation. Based on the nameplate, check whether the ordered device is permitted for the
intended use in the hazardous area (e.g. explosion protection, pressure vessel safety). Use the measuring device only for media to which the process-wetted materials are
sufficiently resistant. If the ambient temperature of the measuring device is outside the atmospheric
temperature, it is absolutely essential to comply with the relevant basic conditions as specified in the device documentation→  8. Protect the measuring device permanently against corrosion from environmental
influences.
Incorrect use
Non-designated use can compromise safety. The manufacturer is not liable for damage caused by improper or non-designated use.
WARNING
L
Danger of breakage due to corrosive or abrasive fluids and ambient conditions!
Verify the compatibility of the process fluid with the sensor material.
Ensure the resistance of all fluid-wetted materials in the process.
Keep within the specified pressure and temperature range.
Endress+Hauser 9
Safety instructions Proline Prowirl F 200 HART
NOTICE
Verification for borderline cases:
For special fluids and fluids for cleaning, Endress+Hauser is glad to provide assistance
in verifying the corrosion resistance of fluid-wetted materials, but does not accept any warranty or liability as minute changes in the temperature, concentration or level of contamination in the process can alter the corrosion resistance properties.
Residual risks
WARNING
L
The electronics and the medium may cause the surfaces to heat up. This presents a burn hazard!
For elevated fluid temperatures, ensure protection against contact to prevent burns.

2.3 Workplace safety

For work on and with the device:
Wear the required personal protective equipment according to federal/national
regulations.
For welding work on the piping:
Do not ground the welding unit via the measuring device.
If working on and with the device with wet hands:
Due to the increased risk of electric shock, gloves must be worn.

2.4 Operational safety

Risk of injury.
Operate the device in proper technical condition and fail-safe condition only.
The operator is responsible for interference-free operation of the device.
Conversions to the device
Unauthorized modifications to the device are not permitted and can lead to unforeseeable dangers.
If, despite this, modifications are required, consult with Endress+Hauser.
Repair
To ensure continued operational safety and reliability,
Carry out repairs on the device only if they are expressly permitted.
Observe federal/national regulations pertaining to repair of an electrical device.
Use original spare parts and accessories from Endress+Hauser only.

2.5 Product safety

This measuring device is designed in accordance with good engineering practice to meet state-of-the-art safety requirements, has been tested, and left the factory in a condition in which it is safe to operate.
It meets general safety standards and legal requirements. It also complies with the EU directives listed in the device-specific EU Declaration of Conformity. Endress+Hauser confirms this by affixing the CE mark to the device.
10 Endress+Hauser
Proline Prowirl F 200 HART Safety instructions

2.6 IT security

Our warranty is valid only if the device is installed and used as described in the Operating Instructions. The device is equipped with security mechanisms to protect it against any inadvertent changes to the settings.
IT security measures, which provide additional protection for the device and associated data transfer, must be implemented by the operators themselves in line with their security standards.

2.7 Device-specific IT security

The device offers a range of specific functions to support protective measures on the operator's side. These functions can be configured by the user and guarantee greater in­operation safety if used correctly. An overview of the most important functions is provided in the following section.

2.7.1 Protecting access via hardware write protection

Write access to the device parameters via the local display or operating tool (e.g. FieldCare, DeviceCare) can be disabled via a write protection switch (DIP switch on the motherboard). When hardware write protection is enabled, only read access to the parameters is possible.

2.7.2 Protecting access via a password

A password can be used to protect against write access to the device parameters.
This password locks write access to the device parameters via the local display or another operating tool (e.g. FieldCare, DeviceCare) and, in terms of functionality, is equivalent to hardware write protection. If the service interface CDI RJ-45 is used, read access is only possible if the password is entered.
User-specific access code
Write access to the device parameters via the local display or operating tool (e.g. FieldCare, DeviceCare) can be protected by the modifiable, user-specific access code (→  124).
When the device is delivered, the device does not have an access code and is equivalent to 0000 (open).
General notes on the use of passwords
• The access code and network key supplied with the device should be changed during commissioning.
• Follow the general rules for generating a secure password when defining and managing the access code or network key.
• The user is responsible for the management and careful handling of the access code and network key.
• For information on configuring the access code or on what to do if you lose the password, see the "Write protection via access code" section →  124

2.7.3 Access via fieldbus

Cyclic fieldbus communication (read and write, e.g. measured value transmission) with a higher-order system is not affected by the restrictions mentioned above.
Endress+Hauser 11
Product description Proline Prowirl F 200 HART
1 2
3
4
5 6
7
8
9
+
E

3 Product description

The device consists of a transmitter and a sensor.
Two device versions are available:
• Compact version – transmitter and sensor form a mechanical unit.
• Remote version - transmitter and sensor are mounted in separate locations.

3.1 Product design

A0020649
 1 Important components of a measuring device
1 Electronics compartment cover 2 Display module 3 Main electronics module 4 Cable glands 5 Transmitter housing (incl. HistoROM) 6 I/O electronics module 7 Terminals (spring loaded terminals, pluggable) 8 Connection compartment cover 9 Sensor
12 Endress+Hauser
Proline Prowirl F 200 HART Product description
1
2
A0034152
 2 Versions of pressure measuring unit
1 Order code for "Sensor version; DSC sensor; measuring tube", option DA "Mass steam" 2 Order code for "Sensor version; DSC sensor; measuring tube", option DB "Mass gas/liquid"
For order code for "Sensor version; DSC sensor; measuring tube", option DA "Mass steam" and DB "Mass gas/liquid", the following applies: Oil-free or grease-free cleaning is not possible
Endress+Hauser 13

Incoming acceptance and product identification Proline Prowirl F 200 HART

1 2
1 2
Order code: Ser. no.: Ext. ord. cd.:
i
i
Date:
4 Incoming acceptance and product
identification

4.1 Incoming acceptance

Are the order codes on the delivery note (1) and the product sticker (2) identical?
Are the goods undamaged?
Do the nameplate data match the ordering information on the delivery note?
Is the envelope present with accompanying documents?
• If one of the conditions is not satisfied, contact your Endress+Hauser Sales Center.
• Depending on the device version, the CD-ROM might not be part of the delivery! The Technical Documentation is available via the Internet or via the Endress+Hauser Operations App, see the "Product identification" section →  15.

4.2 Product identification

The following options are available for identification of the device:
• Nameplate specifications
• Order code with breakdown of the device features on the delivery note
• Enter serial numbers from nameplates in the W@M Device Viewer (www.endress.com/deviceviewer): All information about the device is displayed.
• Enter the serial number from nameplates in the Endress+Hauser Operations App or scan the 2-D matrix code (QR code) on the nameplate using the Endress+Hauser Operations App: All information about the device is displayed.
14 Endress+Hauser
Proline Prowirl F 200 HART Incoming acceptance and product identification
Order code:
Ext. ord. cd.:
Ser. no.:
Date:
i
i
Patents
322540-0001
1 2
3 4 5
6
7 8
9
10 11
12
14
15
16
17
13
For an overview of the scope of the associated Technical Documentation, refer to the following:
• The "Additional standard documentation on the device"→  8 and "Supplementary device-dependent documentation"→  8 sections
• The W@M Device Viewer: enter the serial number from the nameplate (www.endress.com/deviceviewer)
• The Endress+Hauser Operations App: Enter the serial number from the nameplate or scan the 2-D matrix code (QR code) on the nameplate.

4.2.1 Transmitter nameplate

A0032237
 3 Example of a transmitter nameplate
1 Manufacturing location 2 Name of the transmitter 3 Order code 4 Serial number (ser. no.) 5 Extended order code (Ext. ord. cd.) 6 Electrical connection data, e.g. available inputs and outputs, supply voltage 7 Type of cable glands 8 Permitted ambient temperature (Ta)
9 Firmware version (FW) and device revision (Dev.Rev.) from the factory 10 CE mark, C-Tick 11 Additional information on version: certificates, approvals 12 Permitted temperature range for cable 13 Manufacturing date: year-month 14 Degree of protection 15 Approval information for explosion protection 16 Document number of safety-related supplementary documentation 17 2-D matrix code
Endress+Hauser 15
Incoming acceptance and product identification Proline Prowirl F 200 HART
Ser. no.:
Ptest:
Size:
Materials:
Gasket:
Tm:
Qmax(G):
Ta:
1
2 6
7 8
3
9
10
13 12
4 5
14 11

4.2.2 Sensor nameplate

Order code for "Housing" option B "GT18 dual compartment, 316L, compact" and option K "GT18 dual compartment, 316L, remote"
A0034423
 4 Example of a sensor nameplate
1 Name of the sensor 2 Nominal diameter of sensor 3 Flange nominal diameter/nominal pressure 4 Serial number (ser. no.) 5 Measuring tube material 6 Measuring tube material 7 Maximum permitted volume flow (gas/steam): Q
8 Test pressure of the sensor: OPL→  194 9 Seal material 10 Document number of safety-related supplementary documentation →  207 11 Ambient temperature range 12 CE mark 13 Medium temperature range 14 Degree of protection
→  174
max
16 Endress+Hauser
Proline Prowirl F 200 HART Incoming acceptance and product identification
Ser. no.:
Ptest:
Size:
Materials:
Tm:
Qmax(G):
Gasket:
Ta:
1
2
5 6 8
7
9
3
4
10
111213
Order code for "Housing" option C "GT20 dual compartment, aluminum, coated, compact"
A0034161
 5 Example of a sensor nameplate
1 Nominal diameter of sensor 2 Flange nominal diameter/nominal pressure 3 Measuring tube material 4 Measuring tube material 5 Serial number (ser. no.) 6 Maximal permitted volume flow (gas/steam) 7 Test pressure of the sensor 8 Degree of protection 9 Approval information for explosion protection and Pressure Equipment Directive→  207 10 CE mark 11 Seal material 12 Medium temperature range 13 Ambient temperature range
Endress+Hauser 17
Incoming acceptance and product identification Proline Prowirl F 200 HART
Order code:
Ext. ord. cd.:
Ser. no.:
Ptest:
Size:
Ta:
Materials:
Gasket: Tm:
Qmax(G):
4
1
5
8
12
2
7
9
10 11
13141516
3
6
Order code for "Housing" option J "GT20 dual compartment, aluminum, coated, remote"
A0034162
 6 Example of a sensor nameplate
1 Name of the sensor 2 Nominal diameter of sensor 3 Flange nominal diameter/nominal pressure 4 Order code 5 Serial number (ser. no.) 6 Extended order code (Ext. ord. cd.) 7 Maximal permitted volume flow (gas/steam) 8 Degree of protection 9 Approval information for explosion protection and Pressure Equipment Directive 10 Ambient temperature range 11 Document number of safety-related supplementary documentation →  207 12 Test pressure of the sensor 13 Measuring tube material 14 Measuring tube material 15 Seal material 16 Medium temperature range
Order code
The measuring device is reordered using the order code.
Extended order code
• The device type (product root) and basic specifications (mandatory features) are always listed.
• Of the optional specifications (optional features), only the safety and approval­related specifications are listed (e.g. LA). If other optional specifications are also ordered, these are indicated collectively using the # placeholder symbol (e.g. #LA#).
• If the ordered optional specifications do not include any safety and approval-related specifications, they are indicated by the + placeholder symbol (e.g. XXXXXX-ABCDE +).
18 Endress+Hauser
Proline Prowirl F 200 HART Incoming acceptance and product identification
DCP21
1
2 3 4
5
6
7
8
9
Date:

4.2.3 Pressure measuring cells nameplate

A0034354
 7 Example of pressure measuring cell nameplate
1 Manufacturer address 2 Pressure range 3 Maximum permitted pressure 4 Ambient temperature range 5 Serial number or XPD structure 6 Degree of protection 7 CE mark, C-Tick mark 8 QR code 9 Manufacturing date

4.2.4 Symbols on measuring device

Symbol Meaning
WARNING!
This symbol alerts you to a dangerous situation. Failure to avoid this situation can result in serious or fatal injury.
Reference to documentation
Refers to the corresponding device documentation.
Protective ground connection
A terminal which must be connected to ground prior to establishing any other connections.
Endress+Hauser 19
Storage and transport Proline Prowirl F 200 HART

5 Storage and transport

5.1 Storage conditions

Observe the following notes for storage:
Store in the original packaging to ensure protection from shock.
Do not remove protective covers or protective caps installed on process connections.
They prevent mechanical damage to the sealing surfaces and contamination in the measuring tube. Protect from direct sunlight to avoid unacceptably high surface temperatures.
Store in a dry and dust-free place.
Do not store outdoors.
Storage temperature: –50 to +80 °C (–58 to +176 °F)

5.2 Transporting the product

Transport the measuring device to the measuring point in the original packaging.
A0029252
Do not remove protective covers or caps installed on process connections. They prevent mechanical damage to the sealing surfaces and contamination in the measuring tube.

5.2.1 Measuring devices without lifting lugs

WARNING
L
Center of gravity of the measuring device is higher than the suspension points of the webbing slings.
Risk of injury if the measuring device slips.
Secure the measuring device against slipping or turning.
Observe the weight specified on the packaging (stick-on label).
A0029214
20 Endress+Hauser
Proline Prowirl F 200 HART Storage and transport

5.2.2 Measuring devices with lifting lugs

CAUTION
L
Special transportation instructions for devices with lifting lugs
Only use the lifting lugs fitted on the device or flanges to transport the device.
The device must always be secured at two lifting lugs at least.

5.2.3 Transporting with a fork lift

If transporting in wood crates, the floor structure enables the crates to be lifted lengthwise or at both sides using a forklift.

5.3 Packaging disposal

All packaging materials are environmentally friendly and 100 % recyclable:
• Outer packaging of device Polymer stretch wrap that complies with EU Directive 2002/95/EC (RoHS)
• Packaging
• Wooden crate treated in accordance with ISPM 15 standard, confirmed by IPPC logo
• Cardboard box in accordance with European packaging guideline 94/62EC,
recyclability confirmed by Resy symbol
• Carrying and securing materials
• Disposable plastic pallet
• Plastic straps
• Plastic adhesive strips
• Filler material Paper pads
Endress+Hauser 21
Installation Proline Prowirl F 200 HART

6 Installation

6.1 Installation conditions

6.1.1 Mounting position

Mounting location
Orientation
The direction of the arrow on the sensor nameplate helps you to install the sensor according to the flow direction (direction of medium flow through the piping).
Vortex meters require a fully developed flow profile as a prerequisite for correct volume flow measurement. Therefore, please note the following:
Orientation Compact version Remote version
A Vertical orientation
A0015545
B Horizontal orientation, transmitter head up
A0015589
C Horizontal orientation, transmitter head down
A0015590
D Horizontal orientation, transmitter head at side
A0015592




1)
2) 3)
4) 5)
4)




A0015543
1) In the case of liquids, there should be upward flow in vertical pipes to avoid partial pipe filling (Fig. A). Disruption in flow measurement! In the case of vertical orientation and downward flowing liquid, the pipe always needs to be completely filled to ensure correct liquid flow measurement.
2) Danger of electronics overheating! If the fluid temperature is ≥ 200 °C (392 °F), orientation B is not permitted for the wafer version (Prowirl D) with nominal diameters of DN 100 (4") and DN 150 (6").
3) In the case of hot media (e.g. steam or fluid temperature (TM) ≥ 200 °C (392 °F): orientation C or D
4) In the case of very cold media (e.g. liquid nitrogen): orientation B or D
5) For "wet steam detection/measurement" option: orientation C
The "mass" sensor version (integrated pressure/temperature measurement) is available only for measuring devices in the HART communication mode.
22 Endress+Hauser
Proline Prowirl F 200 HART Installation
A
L
Pressure measuring cell
Steam pressure measurement Option DA
E • With the
transmitter installed at the bottom or at the side
• Protection against rising
F
heat
• Reduction in temperature to almost ambient temperature due to siphon
1)
Gas pressure measurement Option DB
G • Pressure
measuring cell with shutoff device above tapping point
• Discharge of any condensate into the process
Liquid pressure measurement Option DB
H Device with shutoff
device at the same level as tapping point

A0034057

A0034058

A0034092

A0034091
1) Note max. permitted ambient temperature of transmitter→  27.
Minimum spacing and cable length
Order code for "Sensor version", option "mass" DA, DB
The "mass" sensor version (integrated pressure/temperature measurement) is available only for measuring devices in the HART communication mode.
A Minimum spacing in all directions L Required cable length
A0019211
Endress+Hauser 23
Installation Proline Prowirl F 200 HART
The following dimensions must be observed to guarantee problem-free access to the device for service purposes:
• A =100 mm (3.94 in)
• L = L + 150 mm (5.91 in)
Inlet and outlet runs
To attain the specified level of accuracy of the measuring device, the inlet and outlet runs mentioned below must be maintained at the very minimum.
24 Endress+Hauser
Proline Prowirl F 200 HART Installation
1
15 DN×
5 × DN
3
25 × DN
5 × DN
40 × DN
5 × DN
4
2
20 × DN
5 × DN
5
20 × DN
5 × DN
6
17 × DN + 8 × h
5 × DN
h
7
50 × DN
5 × DN
9
40 × DN
5 × DN
8
5 × DN
DN 25 (1"):
DN 40 (1½"):
A0019189
 8 Minimum inlet and outlet runs with various flow obstructions
h Difference in expansion 1 Reduction by one nominal diameter size 2 Single elbow (90° elbow) 3 Double elbow (2 × 90° elbows, opposite) 4 Double elbow 3D (2 × 90° elbows, opposite, not on one plane) 5 T-piece 6 Expansion 7 Control valve 8 Two measuring devices in a row where DN ≤ 25 (1"): directly flange on flange 9 Two measuring devices in a row where DN ≥ 40 (1½"): for spacing, see graphic
• If there are several flow disturbances present, the longest specified inlet run must be maintained.
• If the required inlet runs cannot be observed, it is possible to install a specially designed flow conditioner →  25.
The inlet run correction function:
• Makes it possible to shorten the inlet run to a minimum length of 10 × DN in the event of flow obstructions 1 to 4. An additional measuring uncertainty of ±0.5% o.r. occurs here. →  111
• Cannot be combined with the wet steam detection/measurement application package. If wet steam detection/measurement is used, the corresponding inlet runs must be taken into consideration. It is not possible to use a flow conditioner for wet steam.
Flow conditioner
If the inlet runs cannot be observed, the use of a flow conditioner is recommended.
Endress+Hauser 25
Installation Proline Prowirl F 200 HART
8 DN×
2 DN×
5 DN×
1
3…5 × DN
4…8 × DN
PT
TT
The flow conditioner is fitted between two pipe flanges and centered by the mounting bolts. Generally this reduces the inlet run needed to 10 × DN with full accuracy.
A0019208
1 Flow conditioner
The pressure loss for flow conditioners is calculated as follows: ∆ p [mbar] = 0.0085 ⋅ ρ [kg/m3] ⋅ v2 [m/s]
Example for steam Example for H2O condensate (80 °C)
p = 10 bar abs. ρ = 965 kg/m
t = 240 °C → ρ = 4.39 kg/m
v = 40 m/s ∆ p = 0.0085 ⋅ 965 ⋅ 2.5 2 = 51.3 mbar
∆ p = 0.0085 ⋅ 4.394.39 ⋅ 40 2 = 59.7 mbar
3
v = 2.5 m/s
3
ρ : density of the process medium v: average flow velocity abs. = absolute
For the dimensions of the flow conditioner, see the "Technical Information" document, "Mechanical construction" section
Outlet runs when installing external devices
If installing an external device, observe the specified distance.
PT Pressure TT Temperature device
A0019205
Installation dimensions
For the dimensions and installation lengths of the device, see the "Technical Information" document, "Mechanical construction" section.
26 Endress+Hauser
Proline Prowirl F 200 HART Installation

6.1.2 Environment and process requirements

Ambient temperature range
Compact version
Measuring device Non-hazardous area: –40 to +80 °C (–40 to +176 °F)
Ex i, Ex nA, Ex ec: –40 to +70 °C (–40 to +158 °F)
Ex d, XP: –40 to +60 °C (–40 to +140 °F)
Ex d, Ex ia: –40 to +60 °C (–40 to +140 °F)
Local display –40 to +70 °C (–40 to +158 °F)
1) Additionally available as order code for "Test, certificate", option JN "Transmitter ambient temperature – 50 °C (–58 °F)".
2) At temperatures < –20 °C (–4 °F), depending on the physical characteristics involved, it may no longer be possible to read the liquid crystal display.
Remote version
Transmitter Non-hazardous area: –40 to +80 °C (–40 to +176 °F)
Ex i, Ex nA, Ex ec: –40 to +80 °C (–40 to +176 °F)
Ex d: –40 to +60 °C (–40 to +140 °F)
Ex d, Ex ia: –40 to +60 °C (–40 to +140 °F)
Sensor Non-hazardous area: –40 to +85 °C (–40 to +185 °F)
Ex i, Ex nA, Ex ec: –40 to +85 °C (–40 to +185 °F)
Ex d: –40 to +85 °C (–40 to +185 °F)
Ex d, Ex ia: –40 to +85 °C (–40 to +185 °F)
Local display –40 to +70 °C (–40 to +158 °F)
1)
1)
1)
1)
2) 1)
1)
1)
1)
1)
1)
1)
1)
1)
2) 1)
1) Additionally available as order code for "Test, certificate", option JN "Transmitter ambient temperature – 50 °C (–58 °F)".
2) At temperatures < –20 °C (–4 °F), depending on the physical characteristics involved, it may no longer be possible to read the liquid crystal display.
If operating outdoors:
Avoid direct sunlight, particularly in warm climatic regions.
You can order a weather protection cover from Endress+Hauser. →  169.
Thermal insulation
For optimum temperature measurement and mass calculation, heat transfer at the sensor must be avoided for some fluids. This can be ensured by installing thermal insulation. A wide range of materials can be used for the required insulation.
This applies for:
• Compact version
• Remote sensor version
The maximum insulation height permitted is illustrated in the diagram:
Endress+Hauser 27
Installation Proline Prowirl F 200 HART
1
A0019212
1 Maximum insulation height
When insulating, ensure that a sufficiently large area of the housing support remains
exposed.
The uncovered part serves as a radiator and protects the electronics from overheating and excessive cooling.
NOTICE
Electronics overheating on account of thermal insulation!
Observe the maximum permitted insulation height of the transmitter neck so that the
transmitter head and/or the connection housing of the remote version is completely free. Observe information on the permissible temperature ranges.
Note that a certain orientation might be required, depending on the fluid temperature.

6.1.3 Special mounting instructions

Installation for delta heat measurements
• Order code for "Sensor version", option CA "mass; 316L; 316L (integrated temperature measurement), –200 to +400 °C (–328 to +750 °F)"
• Order code for "Sensor version", option CB "mass; Alloy C22; 316L (integrated temperature measurement), –200 to +400 °C (–328 to +750 °F)"
• Order code for "Sensor version", option CC "mass; Alloy C22; Alloy C22 (integrated temperature measurement), –40 to +260 °C (–40 to +500 °F)"
• Order code for "Sensor version", option DA "mass steam; 316L; 316L (integrated pressure/temperature measurement), –200 to +400 °C (–328 to +750 °F)"
• Order code for "Sensor version", option DB "mass gas/liquid; 316L; 316L (integrated pressure/temperature measurement), –40 to +100 °C (–40 to +212 °F)"
The second temperature measurement is taken using a separate temperature sensor. The measuring device reads in this value via a communication interface.
• In the case of saturated steam delta heat measurements, the measuring device must be installed on the steam side.
• In the case of water delta heat measurements, the device can be installed on the cold or warm side.
28 Endress+Hauser
Proline Prowirl F 200 HART Installation
Q
TT
1
2
3
A0019209
 9 Layout for delta heat measurement of saturated steam and water
1 Measuring device 2 Temperature sensor 3 Heat exchanger Q Heat flow
Protective cover
Observe the following minimum head clearance: 222 mm (8.74 in)
For information on the weather protection cover, see →  169

6.2 Mounting the measuring device

6.2.1 Required tools

For transmitter
• For turning the transmitter housing: Open-ended wrench8 mm
• For opening the securing clamps: Allen key3 mm
For sensor
For flanges and other process connections: Corresponding mounting tools

6.2.2 Preparing the measuring device

1. Remove all remaining transport packaging.
2. Remove any protective covers or protective caps present from the sensor.
3. Remove stick-on label on the electronics compartment cover.

6.2.3 Mounting the sensor

WARNING
L
Danger due to improper process sealing!
Ensure that the inside diameters of the gaskets are greater than or equal to that of the
process connections and piping. Ensure that the gaskets are clean and undamaged.
Install the gaskets correctly.
1. Ensure that the direction of the arrow on the sensor matches the flow direction of
Endress+Hauser 29
the medium.
Installation Proline Prowirl F 200 HART
1.
4
2.
1
3
2
4 x
6 mm
2. To ensure compliance with device specifications, install the measuring device between the pipe flanges in a way that it is centered in the measurement section.
3. Install the measuring device or turn the transmitter housing so that the cable entries do not point upwards.
A0029263

6.2.4 Mounting the pressure measuring unit

Preparation
1. Prior to mounting the pressure measuring unit, install the measuring device in the pipe.
2. When mounting the pressure measuring unit, use only the seal provided. The use of a different sealing material is not permitted.
Removing the blind flange
A0034355
1 Mounting screws 2 blind flange 3 Seal 4 Flange connection on sensor side
NOTICE
When replacing the seal following commissioning, fluid may escape when the flange connection is opened!
Ensure that the measuring device is not under pressure.
Ensure that there is no fluid in the measuring device.
1. Release the mounting screws on the blind flange.  The screws are needed again to mount the pressure measuring unit.
30 Endress+Hauser
2. Remove the internal seal.
Loading...