A grounded terminal which, as far as the operator is concerned, is grounded via a
grounding system.
Protective Earth (PE)
A terminal which must be connected to ground prior to establishing any other
connections.
The ground terminals are situated inside and outside the device:
• Inner ground terminal: Connects the protectiv earth to the mains supply.
• Outer ground terminal: Connects the device to the plant grounding system.
Symbols for certain types of information
SymbolMeaning
Permitted
Procedures, processes or actions that are permitted.
Preferred
Procedures, processes or actions that are preferred.
Forbidden
Procedures, processes or actions that are forbidden.
Tip
Indicates additional information.
Reference to documentation.
Reference to page.
Reference to graphic.
Visual inspection.
Symbols in graphics
SymbolMeaning
1, 2, 3, ...Item numbers
, 2., 3., …Series of steps
A, B, C, ...Views
A-A, B-B, C-C, ...Sections
Hazardous area
Safe area (non-hazardous area)
Flow direction
4Endress+Hauser
Page 5
Proline Promass F 100
213
Function and system design
Measuring principle
The measuring principle is based on the controlled generation of Coriolis forces. These forces are
always present in a system when both translational and rotational movements are superimposed.
Fc = 2 · ∆m (ν · ω)
Fc = Coriolis force
∆m = moving mass
ω = rotational velocity
ν = radial velocity in rotating or oscillating system
The amplitude of the Coriolis force depends on the moving mass ∆m, its velocity ν in the system and
thus on the mass flow. Instead of a constant rotational velocity ω, the sensor uses oscillation.
In the sensor, two parallel measuring tubes containing flowing fluid oscillate in antiphase, acting like
a tuning fork. The Coriolis forces produced at the measuring tubes cause a phase shift in the tube
oscillations (see illustration):
• At zero flow (when the fluid is at a standstill) the two tubes oscillate in phase (1).
• Mass flow causes deceleration of the oscillation at the inlet of the tubes (2) and acceleration at the
outlet (3).
A0028850
The phase difference (A-B) increases with increasing mass flow. Electrodynamic sensors register the
tube oscillations at the inlet and outlet. System balance is ensured by the antiphase oscillation of the
two measuring tubes. The measuring principle operates independently of temperature, pressure,
viscosity, conductivity and flow profile.
Density measurement
The measuring tube is continuously excited at its resonance frequency. A change in the mass and
thus the density of the oscillating system (comprising measuring tube and fluid) results in a
corresponding, automatic adjustment in the oscillation frequency. Resonance frequency is thus a
function of medium density. The microprocessor utilizes this relationship to obtain a density signal.
Volume measurement
Together with the measured mass flow, this is used to calculate the volume flow.
Temperature measurement
The temperature of the measuring tube is determined in order to calculate the compensation factor
due to temperature effects. This signal corresponds to the process temperature and is also available
as an output signal.
Endress+Hauser5
Page 6
Proline Promass F 100
Measuring system
The device consists of a transmitter and a sensor. If a device with Modbus RS485 intrinsically safe is
ordered, the Safety Barrier Promass 100 is part of the scope of supply and must be implemented to
operate the device.
The device is available as a compact version:
The transmitter and sensor form a mechanical unit.
• Compact, hygienic, stainless:
– Hygienic version, stainless steel 1.4301 (304)
– Optional: order code for "Sensor feature", option CC
A0016693
A0016694
A0016695
Hygienic version, for maximum corrosion resistance: stainless steel
1.4404 (316L)
• Ultra-compact, hygienic, stainless:
– Hygienic version, stainless steel 1.4301 (304)
– Optional: order code for "Sensor feature", option CC
Hygienic version, for maximum corrosion resistance: stainless steel
1.4404 (316L)
Configuration:
• Via operating tools (e.g. FieldCare, DeviceCare)
• Additionally for device version with local display:
Via Web browser (e.g. Microsoft Internet Explorer)
• Also for device version with 4-20 mA HART, pulse/frequency/switch
output:
Via Web browser (e.g. Microsoft Internet Explorer)
• Also for device version with EtherNet/IP output:
– Via Web browser (e.g. Microsoft Internet Explorer)
– Via Add-on Profile Level 3 for automation system from Rockwell
Automation
– Via Electronic Data Sheet (EDS)
• Also for device version with PROFINET output:
– Via Web browser (e.g. Microsoft Internet Explorer)
– Via device master file (GSD)
Sensor
PromassF• Excellent performance across a wide range of applications
• Simultaneous measurement of flow, volume flow, density and
temperature (multivariable)
• Immune to process influences
• Nominal diameter range: DN 8 to 250 (³⁄₈ to 10")
– Process connections: stainless steel, 1.4404 (316/316L); 1.4301
(304); Alloy C22, 2.4602 (UNS N06022)
Safety Barrier Promass 100
• Dual-channel safety barrier for installation in non-hazardous locations
or zone 2/div. 2:
– Channel 1: DC 24 V power supply
– Channel 2: Modbus RS485
• In addition to current, voltage and power limitation, it offers galvanic
isolation of circuits for explosion protection.
• Easy top-hat rail mounting (DIN 35 mm) for installation in control
A0016763
cabinets
6Endress+Hauser
Page 7
Proline Promass F 100
23
7
8
6
9
10
11
4
1
5
Equipment architecture
A0016779
1Possibilities for integrating measuring devices into a system
1Control system (e.g. PLC)
2EtherNet/IP
3PROFIBUS DP
4PROFINET
5Modbus RS485
64-20 mA HART, pulse/frequency/switch output
7Safety Barrier Promass 100
8Modbus RS485 intrinsically safe
9Non-hazardous area
10 Non-hazardous area and Zone 2/Div. 2
11 Hazardous area and Zone 1/Div. 1
SafetyIT security
We only provide a warranty if the device is installed and used as described in the Operating
Instructions. The device is equipped with security mechanisms to protect it against any inadvertent
changes to the device settings.
IT security measures in line with operators' security standards and designed to provide additional
protection for the device and device data transfer must be implemented by the operators themselves.
Endress+Hauser7
Page 8
Input
Measured variableDirect measured variables
• Mass flow
• Density
• Temperature
Calculated measured variables
• Volume flow
• Corrected volume flow
• Reference density
Measuring rangeMeasuring ranges for liquids
Proline Promass F 100
DNMeasuring range full scale values
min(F)
to
max(F)
[mm][in][kg/h][lb/min]
8³⁄₈0 to 20000 to 73.50
15½0 to 65000 to 238.9
2510 to 180000 to 661.5
401½0 to 450000 to 1654
5020 to 700000 to 2573
8030 to 1800000 to 6615
10040 to 3500000 to 12860
15060 to 8000000 to 29400
250100 to 22000000 to 80850
Measuring ranges for gases
The full scale values depend on the density of the gas and can be calculated with the formula below:
=
max(G)
max(G)
max(F)
<
max(G)
ρ
G
xConstant dependent on nominal diameter
max(F)
max(F)
· ρG : x
Maximum full scale value for gas [kg/h]
Maximum full scale value for liquid [kg/h]
can never be greater than
max(G)
max(F)
Gas density in [kg/m³] at operating conditions
DNx
[mm][in][kg/m3]
8³⁄₈60
15½80
25190
401½90
50290
803110
1004130
8Endress+Hauser
Page 9
Proline Promass F 100
DNx
[mm][in][kg/m3]
1506200
25010200
To calculate the measuring range, use the Applicator sizing tool → 100
Calculation example for gas
• Sensor: Promass F, DN 50
• Gas: Air with a density of 60.3 kg/m³ (at 20 °C and 50 bar)
• Measuring range (liquid): 70000 kg/h
• x = 90 kg/m³ (for Promass F, DN 50)
Maximum possible full scale value:
• Option 85: ATEX II2G + IECEx Z1 Ex ia + CSA C/US
IS Cl. I, II, III Div. 1
Low flow cut off
Protocol-specific dataHART
The switch points for low flow cut off are user-selectable.
Manufacturer ID0x11
Device type ID0x4A
HART protocol revision7
Device description files
(DTM, DD)
HART loadMin. 250 Ω
Order code for
"Approval"
For an overview and for information on the interdependencies between the gas group - sensor - nominal
diameter, see the "Safety Instructions" (XA) document for the measuring device
Information and files under:
www.endress.com
Supply voltageSignal transmission
20 (L-)10 (L+)62 (A)72 (B)
Terminal numbers
Ui = 16.24 V
Ii = 623 mA
Pi = 2.45 W
Li = 0 µH
Ci = 6 nF
Endress+Hauser13
Page 14
Dynamic variablesRead out the dynamic variables: HART command 3
The measured variables can be freely assigned to the dynamic variables.
Measured variables for PV (primary dynamic variable)
• Mass flow
• Volume flow
• Corrected volume flow
• Density
• Reference density
• Temperature
Measured variables for SV, TV, QV (secondary, tertiary and quaternary
dynamic variable)
• Mass flow
• Volume flow
• Corrected volume flow
• Density
• Reference density
• Temperature
• Totalizer 1
• Totalizer 2
• Totalizer 3
The range of options increases if the measuring device has one or more
application packages.
Heartbeat Technology application package
Additional measured variables are available with the Heartbeat Technology
application package:
• Carrier pipe temperature
• Oscillation amplitude 0
Device variablesRead out the device variables: HART command 9
The device variables are permanently assigned.
A maximum of 8 device variables can be transmitted:
• 0 = mass flow
• 1 = volume flow
• 2 = corrected volume flow
• 3 = density
• 4 = reference density
• 5 = temperature
• 6 = totalizer 1
• 7 = totalizer 2
• 8 = totalizer 3
• 13 = target mass flow
• 14 = carrier mass flow
• 15 = concentration
Proline Promass F 100
PROFIBUS DP
Manufacturer ID0x11
Ident number0x1561
Profile version3.02
Device description files (GSD,
DTM, DD)
Information and files under:
• www.endress.com
On the product page for the device: Documents/Software → Device drivers
• www.profibus.org
14Endress+Hauser
Page 15
Proline Promass F 100
Output values
(from measuring device to
automation system)
Input values
(from automation system to
measuring device)
Supported functions• Identification & Maintenance
Configuration of the device
address
Analog input 1 to 8
• Mass flow
• Volume flow
• Corrected volume flow
• Target mass flow
• Carrier mass flow
• Density
• Reference density
• Concentration
• Temperature
• Carrier pipe temperature
• Electronic temperature
• Oscillation frequency
• Oscillation amplitude
• Frequency fluctuation
• Oscillation damping
• Tube damping fluctuation
• Signal asymmetry
• Exciter current
Digital input 1 to 2
• Partially filled pipe detection
• Low flow cut off
Totalizer 1 to 3
• Mass flow
• Volume flow
• Corrected volume flow
Analog output 1 to 3 (fixed assignment)
• Pressure
• Temperature
• Reference density
Digital output 1 to 3 (fixed assignment)
• Digital output 1: switch positive zero return on/off
• Digital output 2: perform zero point adjustment
• Digital output 3: switch switch output on/off
Totalizer 1 to 3
• Totalize
• Reset and hold
• Preset and hold
• Stop
• Operating mode configuration:
– Net flow total
– Forward flow total
– Reverse flow total
Simplest device identification on the part of the control system and
nameplate
• PROFIBUS upload/download
Reading and writing parameters is up to ten times faster with PROFIBUS
upload/download
• Condensed status
Simplest and self-explanatory diagnostic information by categorizing
diagnostic messages that occur
• System units
– Mass flow
– Mass
– Volume flow
– Volume
– Corrected volume flow
– Corrected volume
– Density
– Reference density
– Temperature
– Pressure
• Concentration application package
– Coefficients A0 to A4
– Coefficients B1 to B3
Order code for "Output":
Option M: Modbus RS485, for use in the intrinsically safe area (connection via Safety Barrier Promass 100)
Endress+Hauser27
10 (L+)20 (L-)62 (B)72 (A)
Page 28
Proline Promass F 100
L
L
1
2
+
_
1
2
EtherNet/IP connection version
Order code for "Output", option N
Depending on the housing version, the transmitters can be ordered with terminals or device plugs.
Order code
"Housing"
Options
A, B
Options
A, B, C
Order code for "Housing":
• Option A: compact, coated aluminum
• Option B: compact, hygienic, stainless
• Option C ultra-compact, hygienic, stainless
Connection methods available
Output
Device plug
connectors
→ 32
Device plug
connectors
→ 32
Power
supply
Terminals• Option L: plug M12x1 + thread NPT ½"
Device plug
connectors
→ 32
Possible options for order code
"Electrical connection"
• Option N: plug M12x1 + coupling M20
• Option P: plug M12x1 + thread G ½"
• Option U: plug M12x1 + thread M20
Option Q: 2 x plug M12x1
6EtherNet/IP terminal assignment
1Power supply: DC 24 V
2EtherNet/IP
Order code
"Output"
Option NDC 24 VEtherNet/IP
Order code for "Output":
Option N: EtherNet/IP
A0017054
Terminal number
Power supplyOutput
2 (L-)1 (L+)Device plug M12x1
28Endress+Hauser
Page 29
Proline Promass F 100
L
L
1
2
+
_
1
2
PROFINET connection version
Order code for "Output", option R
Depending on the housing version, the transmitters can be ordered with terminals or device plugs.
Order code
"Housing"
Options
A, B
Options
A, B, C
Order code for "Housing":
• Option A: compact, coated aluminum
• Option B: compact, hygienic, stainless
• Option C ultra-compact, hygienic, stainless
Connection methods available
Output
Device plug
connectors
→ 30
Device plug
connectors
→ 30
Power
supply
Terminals• Option L: plug M12x1 + thread NPT ½"
Device plug
connectors
→ 30
Possible options for order code
"Electrical connection"
• Option N: plug M12x1 + coupling M20
• Option P: plug M12x1 + thread G ½"
• Option U: plug M12x1 + thread M20
Option Q: 2 x plug M12x1
7PROFINET terminal assignment
1Power supply: DC 24 V
2PROFINET
Order code
"Output"
Option RDC 24 VPROFINET
Order code for "Output":
Option R: PROFINET
A0017054
Terminal number
Power supplyOutput
2 (L-)1 (L+)Device plug M12x1
Endress+Hauser29
Page 30
Safety Barrier Promass 100
A
Safe area
1
L+2L
27
26
B A
Power
supply
24V
DC
Modbus
RS485
10207262
L+LB A
Power
supply
Modbus
RS485
Hazardous area
Safety Barrier
Promass 100
Power
Communication
Lift panel for
bus termination
1
2
A
1
2
4
3
5
8Safety Barrier Promass 100 with terminals
1Non-hazardous area, Zone 2, Class I Division 2
2Intrinsically safe area
Proline Promass F 100
A0030220
Pin assignment, device plug
Order codes for the M12x1 connectors, see the "Order code for electrical connection" column:
• 4-20 mA HART, pulse/frequency/switch output → 23
• PROFIBUS DP→ 25
• Modbus RS485 → 26
• EtherNet/IP → 28
• PROFINET→ 29
Supply voltage
For all connection versions except MODBUS RS485 intrinsically safe (device side)
Device plug MODBUS RS485 intrinsically safe with supply voltage → 31
PinAssignment
1L+DC 24 V
2Not assigned
3Not assigned
4L-DC 24 V
A0029042
5Grounding/shielding
CodingPlug/socket
APlug
The following is recommended as a socket:
• Binder, series 763, part no. 79 3440 35 05
• Alternatively: Phoenix part no. 1669767 SAC-5P-M12MS
– With the order code for "Output", option B: 4-20 mA HART, pulse/frequency/switch output
– With the order code for "Output", option N: EtherNet/IP
• When using the device in a hazardous location: Use a suitably certified socket.
30Endress+Hauser
Page 31
Proline Promass F 100
3
2
4
1
5
3
2
4
1
5
1
2
4
3
5
4-20 mA HART with pulse/frequency/switch output
Device plug for signal transmission (device side)
PinAssignment
1+4-20 mA HART (active)
2-4-20 mA HART (active)
3+Pulse/frequency/switch output (passive)
4-Pulse/frequency/switch output (passive)
A0016810
• Recommended plug: Binder, series 763, part no. 79 3439 12 05
• When using the device in a hazardous location, use a suitably certified plug.
5Grounding/shielding
CodingPlug/socket
ASocket
PROFIBUS DP
For use in the non-hazardous area and Zone 2/Div. 2.
Device plug for signal transmission (device side)
PinAssignment
1Not assigned
2APROFIBUS DP
3Not assigned
4BPROFIBUS DP
A0016811
• Recommended plug: Binder, series 763, part no. 79 4449 20 05
• When using the device in a hazardous location, use a suitably certified plug.
5Grounding/shielding
CodingPlug/socket
BSocket
MODBUS RS485
Device plug for signal transmission with supply voltage (device side), MODBUS RS485 (intrinsically
safe)
PinAssignment
1L+Supply voltage, intrinsically safe
A0029042
2A
3B
4L-Supply voltage, intrinsically safe
5Grounding/shielding
CodingPlug/socket
APlug
Modbus RS485 intrinsically safe
• Recommended socket: Binder, series 763, part no. 79 3439 12 05
• When using the device in a hazardous location: Use a suitably certified socket.
Endress+Hauser31
Page 32
Proline Promass F 100
3
2
4
1
5
3
2
4
1
3
2
4
1
Device plug for signal transmission (device side), MODBUS RS485 (not intrinsically safe)
For use in the non-hazardous area and Zone 2/Div. 2.
PinAssignment
1Not assigned
2AModbus RS485
3Not assigned
4BModbus RS485
A0016811
• Recommended plug: Binder, series 763, part no. 79 4449 20 05
• When using the device in a hazardous location, use a suitably certified plug.
EtherNet/IP
Device plug for signal transmission (device side)
5Grounding/shielding
CodingPlug/socket
BSocket
PinAssignment
1+Tx
2+Rx
3-Tx
4-Rx
A0016812
Recommended plug:
• Binder, series 763, part no. 99 3729 810 04
• Phoenix, part no. 1543223 SACC-M12MSD-4Q
• When using the device in a hazardous location, use a suitably certified plug.
CodingPlug/socket
DSocket
PROFINET
Device plug for signal transmission (device side)
PinAssignment
1+TD +
2+RD +
3-TD –
4-RD –
A0016812
CodingPlug/socket
DSocket
Recommended plug:
• Binder, series 763, part no. 99 3729 810 04
• Phoenix, part no. 1543223 SACC-M12MSD-4Q
• When using the device in a hazardous location, use a suitably certified plug.
Supply voltage
The power unit must be tested to ensure it meets safety requirements (e.g. PELV, SELV).
32Endress+Hauser
Page 33
Proline Promass F 100
Transmitter
For device version with communication type:
• HART, PROFIBUS DP, EtherNet/IP: DC 20 to 30 V
• Modbus RS485, device version:
– For use in the non-hazardous area and Zone 2/Div. 2: DC 20 to 30 V
– For use in the intrinsically safe area: power supply via Safety Barrier Promass 100
Promass 100 safety barrier
DC 20 to 30 V
Power consumptionTransmitter
Order code for "Output"
Option B: 4-20 mA HART with pulse/frequency/switch output3.5 W
Option L: PROFIBUS DP3.5 W
Option M Modbus RS485, for use in non-hazardous areas and Zone 2/
Div. 2
Option M: Modbus RS485, for use in intrinsically safe areas2.45 W
Option N: EtherNet/IP3.5 W
Option R: PROFINET3.5 W
Promass 100 safety barrier
Order code for "Output"
Option M: Modbus RS485, for use in intrinsically safe areas4.8 W
Option M Modbus RS485, for use in non-hazardous
areas and Zone 2/Div. 2
Option M: Modbus RS485, for use in intrinsically safe
areas
Option N: EtherNet/IP145 mA18 A (< 0.125 ms)
Option R: PROFINET145 mA18 A (< 0.125 ms)
Maximum
Power consumption
Maximum
Power consumption
Maximum
Current consumption
90 mA10 A (< 0.8 ms)
145 mA16 A (< 0.4 ms)
switch-on current
3.5 W
Maximum
Promass 100 safety barrier
Power supply failure
Order code for "Output"
Option M: Modbus RS485, for use in intrinsically safe
areas
Maximum
Current consumption
230 mA10 A (< 0.8 ms)
Depending on the device version, the configuration is retained in the device memoryor in the
Maximum
switch-on current
pluggable data memory (HistoROM DAT).
Endress+Hauser33
Page 34
Electrical connectionConnecting the transmitter
121234
ABC
4
4...20 mA
5
2
1
3
6
AHousing version: compact, aluminum coated
BHousing version: compact hygienic, stainless
1Cable entry or device plug for signal transmission
2Cable entry or device plug for supply voltage
CHousing version: ultra-compact, hygienic, stainless, M12 device plug
3Device plug for signal transmission
4Device plug for supply voltage
Proline Promass F 100
A0016924
• Terminal assignment→ 23
• Pin assignment, device plug→ 30
In the case of device versions with a connector, the transmitter housing does not need to be
opened to connect the signal cable or power supply cable.
Connection examples
Current output 4 to 20 mA HART
A0029055
9Connection example for 4 to 20 mA HART current output (active)
1Automation system with current input (e.g. PLC)
2Cable shield: the cable shield must be grounded at both ends to comply with EMC requirements; observe cable
specifications → 39
3Connection for HART operating devices → 91
4Resistor for HART communication (≥ 250 Ω): observe maximum load
5Analog display unit: observe maximum load
6Transmitter
34Endress+Hauser
Page 35
Proline Promass F 100
2
3
4...20 mA
41
5
1
2
3
12345
1
2
3
A0028762
10Connection example for 4 to 20 mA HART current output (passive)
1Automation system with current input (e.g. PLC)
2Power supply
3Cable shield: the cable shield must be grounded at both ends to comply with EMC requirements; observe cable
15Connection example for Modbus RS485 intrinsically safe
1Control system (e.g. PLC)
2Cable shield, observe cable specifications
3Safety Barrier Promass 100
4Observe cable specifications
5Non-hazardous area
6Non-hazardous area and Zone 2/Div. 2
7Intrinsically safe area
8Transmitter
18Connection example for HART input (burst mode) via current output (active)
1Cable shield, observe cable specifications
2Resistor for HART communication (≥ 250 Ω): observe maximum load
3Connection for HART operating devices
4Analog display unit
5Transmitter
6Sensor for external measured variable
A0019828
38Endress+Hauser
Page 39
Proline Promass F 100
3
4...20 mA
4
521
+
-
6
4...20 mA
+
7
3
4
52
+
-
-
+
A0019830
19Connection example for HART input (master mode) via current output (active)
1Automation system with current input (e.g. PLC).
Prerequisite: automation system with HART version 6, HART commands 113 and 114 can be processed.
2Cable shield, observe cable specifications
3Resistor for HART communication (≥ 250 Ω): observe maximum load
4Connection for HART operating devices
5Analog display unit
6Transmitter
7Sensor for external measured variable
Potential equalizationRequirements
No special measures for potential equalization are required.
Please consider the following to ensure correct measurement:
• Same electrical potential for the fluid and sensor
• Company-internal grounding concepts
For devices intended for use in hazardous locations, please observe the guidelines in the
Ex documentation (XA).
Terminals
Transmitter
Spring terminals for wire cross-sections0.5 to 2.5 mm2 (20 to 14 AWG)
Promass 100 safety barrier
Plug-in screw terminals for wire cross-sections0.5 to 2.5 mm2 (20 to 14 AWG)
Cable entries
• Cable gland: M20 × 1.5 with cable Ø 6 to 12 mm (0.24 to 0.47 in)
• Thread for cable entry:
– M20
– G ½"
– NPT ½"
Cable specificationPermitted temperature range
• The installation guidelines that apply in the country of installation must be observed.
• The cables must be suitable for the minimum and maximum temperatures to be expected.
Power supply cable
Standard installation cable is sufficient.
Signal cable
Current output 4 to 20 mA HART
A shielded cable is recommended. Observe grounding concept of the plant.
Pulse/frequency/switch output
Standard installation cable is sufficient.
Endress+Hauser39
Page 40
Proline Promass F 100
PROFIBUS DP
The IEC 61158 standard specifies two types of cable (A and B) for the bus line which can be used for
every transmission rate. Cable type A is recommended.
Cable typeA
Characteristic impedance135 to 165 Ω at a measuring frequency of 3 to 20 MHz
Cable capacitance< 30 pF/m
Wire cross-section
Cable typeTwisted pairs
Loop resistance≤110 Ω/km
Signal dampingMax. 9 dB over the entire length of the cable cross-section
ShieldCopper braided shielding or braided shielding with foil shield. When grounding
Modbus RS485
The EIA/TIA-485 standard specifies two types of cable (A and B) for the bus line which can be used
for every transmission rate. Cable type A is recommended.
> 0.34 mm2 (22 AWG)
the cable shield, observe the grounding concept of the plant.
Cable typeA
Characteristic impedance135 to 165 Ω at a measuring frequency of 3 to 20 MHz
Cable capacitance< 30 pF/m
Wire cross-section
Cable typeTwisted pairs
Loop resistance≤110 Ω/km
Signal dampingMax. 9 dB over the entire length of the cable cross-section
ShieldCopper braided shielding or braided shielding with foil shield. When grounding
> 0.34 mm2 (22 AWG)
the cable shield, observe the grounding concept of the plant.
EtherNet/IP
The standard ANSI/TIA/EIA-568-B.2 Annex specifies CAT 5 as the minimum category for a cable
used for EtherNet/IP. CAT 5e and CAT 6 are recommended.
For more information on planning and installing EtherNet/IP networks, please refer to the
"Media Planning and Installation Manual. EtherNet/IP" of ODVA Organization
PROFINET
Standard IEC 61156-6 specifies CAT 5 as the minimum category for a cable used for PROFINET. CAT
5e and CAT 6 are recommended.
For more information on planning and installing PROFINET networks, see: "PROFINET Cabling
and Interconnection Technology", Guideline for PROFINET
Connecting cable between Safety Barrier Promass 100 and measuring device
Cable typeShielded twisted-pair cable with 2x2 wires. When grounding the cable shield,
observe the grounding concept of the plant.
Maximum cable resistance 2.5 Ω, one side
Comply with the maximum cable resistance specifications to ensure the operational reliability
of the measuring device.
40Endress+Hauser
Page 41
Proline Promass F 100
The maximum cable length for individual wire cross-sections is specified in the table below. Observe
the maximum capacitance and inductance per unit length of the cable and connection values for
hazardous areas .
Wire cross-sectionMaximum cable length
[mm2][AWG][m][ft]
0.52070230
0.7518100328
1.017100328
1.516200656
2.514300984
Performance characteristics
reference operating
conditions
Maximum measured error
• Error limits based on ISO 11631
• Water with +15 to +45 °C (+59 to +113 °F) at2 to 6 bar (29 to 87 psi)
• Specifications as per calibration protocol
• Accuracy based on accredited calibration rigs that are traced to ISO 17025.
To obtain measured errors, use the Applicator sizing tool → 100
o.r. = of reading; 1 g/cm³ = 1 kg/l; T = medium temperature
Base accuracy
Design fundamentals → 44
Mass flow and volume flow (liquids)
±0.05 % o.r. (PremiumCal; order code for "Calibration flow", option D, for mass flow)
±0.10 % o.r.
Mass flow (gases)
±0.35 % o.r.
Density (liquids)
Under
reference operating conditions
[g/cm³][g/cm³][g/cm³]
±0.0005±0.01±0.001
Standard density
calibration
1)
Wide-range
Density specification
2) 3)
1)Valid over the entire temperature and density range
2)Valid range for special density calibration: 0 to 2 g/cm³, +5 to +80 °C (+41 to +176 °F)
3)order code for "Application package", option EE "Special density" only in combination with the order code for
"Measuring tube mat., wetted surface", option BB, BF, HA, SA
Temperature
±0.5 °C ± 0.005 · T °C (±0.9 °F ± 0.003 · (T – 32) °F)
Endress+Hauser41
Page 42
Zero point stability
DNZero point stability
[mm][in][kg/h][lb/min]
8³⁄₈0.0300.001
15½0.2000.007
2510.5400.019
401½2.250.083
5023.500.129
8039.00.330
100414.00.514
150632.01.17
2501088.03.23
Flow values
Flow values as turndown parameter depending on nominal diameter.
The output accuracy must be factored into the measured error if analog outputs are used, but
can be ignored for fieldbus outputs (e.g. Modbus RS485, EtherNet/IP).
The outputs have the following base accuracy specifications.
Current output
AccuracyMax. ±5 µA
Pulse/frequency output
o.r. = of reading
AccuracyMax. ±50 ppm o.r. (over the entire ambient temperature range)
Repeatability
Response time
Influence of ambient
temperature
o.r. = of reading; 1 g/cm3 = 1 kg/l; T = medium temperature
Base repeatability
Design fundamentals → 44
Mass flow and volume flow (liquids)
±0.025 % o.r. (PremiumCal, for mass flow)
±0.05 % o.r.
Mass flow (gases)
±0.25 % o.r.
Density (liquids)
±0.00025 g/cm
3
Temperature
±0.25 °C ± 0.0025 · T °C (±0.45 °F ± 0.0015 · (T–32) °F)
The response time depends on the configuration (damping).
Current output
o.r. = of reading
Temperature coefficientMax. ±0.005 % o.r./°C
Pulse/frequency output
Temperature coefficientNo additional effect. Included in accuracy.
Influence of medium
temperature
Mass flow and volume flow
o.f.s. = of full scale value
When there is a difference between the temperature for zero point adjustment and the process
temperature, the additional measured error of the sensor is typically ±0.0002 % o.f.s./°C (±0.0001 %
o. f.s./°F).
The effect is reduced if zero point adjustment is performed at process temperature.
Density
When there is a difference between the density calibration temperature and the process
temperature, the typical measured error of the sensor is
±0.00005 g/cm3 /°C (±0.000025 g/cm3 /°F). Field density calibration is possible.
Endress+Hauser43
Page 44
Proline Promass F 100
3
[°C]
0- 0880 160 240
[° ]F
- 050
100
200
300
0
2
4
6
8
10
2
1
12
14
16
18
320 400 480 560
Wide-range density specification (special density calibration)
If the process temperature is outside the valid range (→ 41) the measured error is
±0.00005 g/cm3 /°C (±0.000025 g/cm3 /°F)
Influence of medium
pressure
Design fundamentals
1Field density calibration, for example at +20 °C (+68 °F)
2Special density calibration
Temperature
±0.005 · T °C (± 0.005 · (T – 32) °F)
The table below shows the effect on accuracy of mass flow due to a difference between calibration
pressure and process pressure.
o.r. = of reading
It is possible to compensate for the effect by:
• Reading in the current pressure measured value via the current input.
• Specifying a fixed value for the pressure in the device parameters.
Operating Instructions.
DN[% o.r./bar][% o.r./psi]
[mm][in]
83/8no influence
15½no influence
251no influence
401½–0.003–0.0002
502–0.008–0.0006
803–0.009–0.0006
1004–0.007–0.0005
1506–0.009–0.0006
25010–0.009–0.0006
o.r. = of reading, o.f.s. = of full scale value
BaseAccu = base accuracy in % o.r., BaseRepeat = base repeatability in % o.r.
MeasValue = measured value; ZeroPoint = zero point stability
A0034654
44Endress+Hauser
Page 45
Proline Promass F 100
ZeroPoint
BaseAccu
⋅ 100
³
± BaseAccu
ZeroPoint
BaseAccu
⋅ 100
<
ZeroPoint
MeasValue
⋅ 100
±
½ ⋅ ZeroPoint
BaseRepeat
⋅ 100
³
½ ⋅ ZeroPoint
BaseRepeat
⋅ 100
<
ZeroPoint
MeasValue
⋅ 100
½ ⋅±
E [%]
0102030405060
0
0.5
1.0
1.5
2.0
2.5
708090100
Q [%]
Calculation of the maximum measured error as a function of the flow rate
Flow rateMaximum measured error in % o.r.
A0021332
A0021333
Calculation of the maximum repeatability as a function of the flow rate
Flow rateMaximum repeatability in % o.r.
A0021335
A0021336
Example for maximum measured error
A0021339
A0021334
A0021340
A0021337
Mounting location
Endress+Hauser45
A0028808
EMaximum measured error in % o.r. (example with PremiumCal)
QFlow rate in % of maximum full scale value
Installation
No special measures such as supports etc. are necessary. External forces are absorbed by the
construction of the device.
A0028772
To prevent measuring errors arising from accumulation of gas bubbles in the measuring tube, avoid
the following mounting locations in the pipe:
• Highest point of a pipeline.
• Directly upstream of a free pipe outlet in a down pipe.
Page 46
Proline Promass F 100
1
2
3
4
5
Installation in down pipes
However, the following installation suggestion allows for installation in an open vertical pipeline.
Pipe restrictions or the use of an orifice with a smaller cross-section than the nominal diameter
prevent the sensor running empty while measurement is in progress.
20Installation in a down pipe (e.g. for batching applications)
1Supply tank
2Sensor
3Orifice plate, pipe restriction
4Valve
5Batching tank
A0028773
Orientation
DNØ orifice plate, pipe restriction
[mm][in][mm][in]
8³⁄₈60.24
15½100.40
251140.55
401½220.87
502281.10
803501.97
1004652.60
1506903.54
250101505.91
The direction of the arrow on the sensor nameplate helps you to install the sensor according to the
flow direction (direction of medium flow through the piping).
OrientationRecommendation
AVertical orientation
BHorizontal orientation, transmitter at
top
A0015591
Exceptions:
→ 21, 47
A0015589
46Endress+Hauser
1)
Page 47
Proline Promass F 100
12
OrientationRecommendation
CHorizontal orientation, transmitter at
bottom
→ 21, 47
A0015590
DHorizontal orientation, transmitter at
side
1)Applications with low process temperatures may decrease the ambient temperature. To maintain the
minimum ambient temperature for the transmitter, this orientation is recommended.
2)Applications with high process temperatures may increase the ambient temperature. To maintain the
maximum ambient temperature for the transmitter, this orientation is recommended.
A0015592
2)
Exceptions:
If a sensor is installed horizontally with a curved measuring tube, match the position of the sensor to
the fluid properties.
Inlet and outlet runs
Special mounting
instructions
A0028774
21Orientation of sensor with curved measuring tube
1Avoid this orientation for fluids with entrained solids: Risk of solids accumulating.
2Avoid this orientation for outgassing fluids: Risk of gas accumulating.
No special precautions need to be taken for fittings which create turbulence, such as valves, elbows
or T-pieces, as long as no cavitation occurs → 57.
Rupture disk
Information that is relevant to the process: → 56.
The position of the rupture disk is indicated on a sticker beside it.
The transportation guard must be removed.
The existing connecting nozzles are not intended for the purpose of rinsing or pressure monitoring,
but instead serve as the mounting location for the rupture disk.
In the event of a failure of the rupture disk, a discharge device can be screwed onto the internal
thread of the rupture disk in order to drain off any escaping medium.
Endress+Hauser47
Page 48
RUPTURE DISK
i
2
3
1
2
3
1Rupture disk label
> 22.5
(> 0.89)
> 22.5
(> 0.89)
2Rupture disk with 1/2" NPT internal thread with 1" width across flat
3Transport protection
Proline Promass F 100
A0028903
Installing the Safety Barrier
Promass 100
For information on the dimensions: see the "Mechanical construction -> Accessories" section
Zero point adjustment
All measuring devices are calibrated in accordance with state-of-the-art technology. Calibration
takes place under reference conditions→ 41. Therefore, a zero point adjustment in the field is
generally not required.
Experience shows that zero point adjustment is advisable only in special cases:
• To achieve maximum measuring accuracy even with low flow rates.
• Under extreme process or operating conditions (e.g. very high process temperatures or very highviscosity fluids).
A0016894
22Minimum distance between additional Safety Barrier Promass 100 or other modules. Engineering unit
mm (in)
48Endress+Hauser
Page 49
Proline Promass F 100
Environment
Ambient temperature range
Storage temperature
Climate class
Degree of protection
Vibration resistance
Measuring device• –40 to +60 °C (–40 to +140 °F)
• Order code for "Test, certificate", option JM:
–50 to +60 °C (–58 to +140 °F)
Safety Barrier Promass 100–40 to +60 °C (–40 to +140 °F)
If operating outdoors:
‣
Avoid direct sunlight, particularly in warm climatic regions.
–40 to +80 °C (–40 to +176 °F), preferably at +20 °C (+68 °F) (standard version)
–50 to +80 °C (–58 to +176 °F) (Order code for "Test, certificate", option JM)
DIN EN 60068-2-38 (test Z/AD)
Transmitter and sensor
• As standard: IP66/67, type 4X enclosure
• With the order code for "Sensor options", option CM: IP69 can also be ordered
• When housing is open: IP20, type 1 enclosure
• Display module: IP20, type 1 enclosure
Safety Barrier Promass 100
IP20
• Oscillation, sinusoidal, following IEC 60068-2-6
– 2 to 8.4 Hz, 3.5 mm peak
– 8.4 to 2000 Hz, 1 g peak
• Oscillation, broadband noise following IEC 60068-2-64
– 10 to 200 Hz, 0.003 g2/Hz
– 200 to 2000 Hz, 0.001 g2/Hz
– Total: 1.54 g rms
Shock resistance
Shock resistance
Interior cleaning
Electromagnetic
compatibility (EMC)
Shock, half-sine according to IEC 60068-2-27
6 ms 30 g
Shock due to rough handling following IEC 60068-2-31
• Cleaning in place (CIP)
• Sterilization in place (SIP)
Options
• Oil- and grease-free version for wetted parts, without declaration
Order code for "Service", option HA
• Oil- and grease-free version for wetted parts as per IEC/TR 60877-2.0 and BOC 50000810-4, with
declaration
Order code for "Service", option HB
• Depends on the communication protocol:
– HART, PROFIBUS DP, EtherNet/IP:
As per IEC/EN 61326 and NAMUR Recommendation 21 (NE 21)
– Modbus RS485:
As per IEC/EN 61326 and NAMUR Recommendation 21 (NE 21)
– PROFINET: as per IEC/EN 61326
• Complies with emission limits for industry as per EN 55011 (Class A)
• Device version with PROFIBUS DP: Complies with emission limits for industry as per EN 50170
Volume 2, IEC 61784
The following applies for PROFIBUS DP: If baud rates > 1.5 MBaud, an EMC cable entry must be
used and the cable shield must continue as far as the terminal wherever possible.
Details are provided in the Declaration of Conformity.
Endress+Hauser49
Page 50
Medium temperature range
T
a
T
m
B
A
Proline Promass F 100
Process
Standard version–50 to +150 °C (–58 to +302 °F)Order code for "Measuring tube
mat., wetted surface", option HA,
SA, SB, SC
Extended temperature version–50 to +240 °C (–58 to +464 °F)Order code for "Measuring tube
mat., wetted surface", option SD,
SE, SF, TH
Dependency of ambient temperature on medium temperature
23Exemplary representation, values in the table below.
TaAmbient temperature
TmMedium temperature
AMaximum permitted medium temperature Tm at T
require a reduced ambient temperature T
a
= 60 °C (140 °F); higher medium temperatures T
a max
BMaximum permitted ambient temperature Ta for the maximum specified medium temperature Tm of the
sensor
Values for devices used in the hazardous area:
Separate Ex documentation (XA) for the device .
Not insulatedInsulated
ABAB
VersionT
a
T
m
T
a
T
m
T
a
T
m
T
a
Standard version60 °C (140 °F) 150 °C (302 °F)––60 °C (140 °F) 110 °C (230 °F) 55 °C (131 °F) 150 °C (302 °F)
Extended
60 °C (140 °F) 170 °C (338 °F) 55 °C (131 °F) 240 °C (464 °F) 60 °C (140 °F) 110 °C (230 °F) 50 °C (122 °F) 240 °C (464 °F)
temperature
version
Density
Pressure-temperature curves
0 to 5000 kg/m3 (0 to 312 lb/cf)
The following pressure/temperature diagrams apply to all pressure-bearing parts of the device and
not just the process connection. The diagrams show the maximum permissible medium pressure
depending on the specific medium temperature.
Pressure-temperature curves with temperature range +151 to +240 °C (+304 to +464 °F)
exclusively for extended temperature version of measuring devices.
A0031121
m
T
m
50Endress+Hauser
Page 51
Proline Promass F 100
PN100
PN63
PN40
PN16
-80
080160
240
320400480
0
10
20
30
40
50
60
70
80
90
100
0
200
400
600
800
[°C]
[°F]
[bar][psi]
25020015010050
0
-50
Class 600
Class 300
Class 150
-80
080160
240
320400480
0
10
20
30
40
50
60
70
80
90
100
0
200
400
600
800
[°C]
[°F]
[bar][psi]
25020015010050
0
-50
Flange according to EN 1092-1 (DIN 2501)
A0034658-EN
24With flange material 1.4404 (F316/F316L), Alloy C22
Flange according to ASME B16.5
A0034659-EN
25With flange material 1.4404 (F316/F316L)
Endress+Hauser51
Page 52
Class 600
Class 300
Class 150
-80
080160
240
320400480
0
10
20
30
40
50
60
70
80
90
100
0
200
400
600
800
[°C]
[°F]
[bar][psi]
25020015010050
0
-50
26With flange material Alloy C22
63K
40K
20K
10K
-80
080160
240
320400480
0
10
20
30
40
50
60
70
0
200
400
600
800
[°C]
[°F]
[bar][psi]
25020015010050
0
-50
DN8…40
≥ DN 50
-80
080160
240
320400480
0
10
20
30
40
0
[°C]
[°F]
[bar][psi]
25020015010050
0
-50
50
Proline Promass F 100
A0034660-EN
Flange JIS B2220
27With flange material 1.4404 (F316/F316L), Alloy C22
Flange DIN 11864-2 Form A
A0034665-EN
28With connection material 1.4404 (316/316L)
52Endress+Hauser
A0028782-EN
Page 53
Proline Promass F 100
PN40
-80
080160
240
320400480
0
10
20
30
40
0
[°C]
[°F]
[bar][psi]
25020015010050
0
-50
50
Class 150
Class 600
Class 300
-80
080160
240
320400480
0
10
20
30
40
50
60
70
80
90
100
0
200
400
600
800
[°C]
[°F]
[bar][psi]
25020015010050
0
-50
20K
-80
080160
240
320400480
0
10
20
30
0
200
400
[°C]
[°F]
[bar]
[psi]
25020015010050
0
-50
Lap joint flange according to EN 1092-1 (DIN 2501)
A0028784-EN
29With flange material 1.4301 (F304); wetted parts Alloy C22
Lap joint flange according to ASME B16.5
A0028785-EN
30With flange material 1.4301 (F304); wetted parts Alloy C22
Lap joint flange JIS B2220
Endress+Hauser53
31With flange material 1.4301 (F304); wetted parts Alloy C22
A0028786-EN
Page 54
Proline Promass F 100
≥ DN 50
DN8…40
-80
080160
240
320400480
0
10
20
30
40
0
[°C]
[°F]
[bar][psi]
25020015010050
0
-50
50
DN8…40
≥ DN 50
-80
080160
240
320400480
0
10
20
30
40
0
[°C]
[°F]
[bar][psi]
25020015010050
0
-50
50
-80
080160
240
320400480
0
10
20
0
[°C]
[°F]
[bar][psi]
25020015010050
0
-50
Thread DIN 11851
A0028794-EN
32With connection material 1.4404 (316/316L)
DIN 11851 allows for applications up to +140 °C (+284 °F) if suitable sealing materials are used.
Please take this into account when selecting seals and counterparts, as these components can limit
the pressure and temperature range.
Thread DIN 11864-1 Form A
33With connection material 1.4404 (316/316L)
Thread ISO 2853
34With connection material 1.4404 (316/316L)
54Endress+Hauser
A0028798-EN
A0028799-EN
Page 55
Proline Promass F 100
-80
080160
240
320400480
0
10
20
0
[°C]
[°F]
[bar][psi]
25020015010050
0
-50
60
70
80
90
100
800
[bar][psi]
-80
080160
240
320400480
[°C]
[°F]
25020015010050
0
-50
-80
080160
240
320400480
0
10
20
0
[°C]
[°F]
[bar][psi]
25020015010050
0
-50
Thread SMS 1145
A0028800-EN
35With connection material 1.4404 (316/316L)
SMS 1145 allows for applications up to 16 bar (232 psi) if suitable sealing materials are used. Please
take this into account when selecting seals and counterparts, as these components can limit the
pressure and temperature range.
VCO
36With connection material 1.4404 (316/316L)
Tri-Clamp
The clamp connections are suitable up to a maximum pressure of 16 bar (232 psi). Please observe
the operating limits of the clamp and seal used as they can be over 16 bar (232 psi). The clamp and
seal are not included in the scope of supply.
Secondary containment
For the Standard version with the temperature range –50 to +150 °C (–58 to +302 °F), the sensor
housing is filled with dry nitrogen gas and protects the electronics and mechanics inside.
Endress+Hauser55
For all other temperature versions the sensor housing is filled with dry inert gas.
The following secondary containment pressure ratings/burst pressures are only valid for standard
devices and/or devices equipped with closed purge connections (never opened/as delivered).
If a device fitted with purge connections (order code for "Sensor option", option CH "Purge
connection") is connected to the purge system, the maximum nominal pressure is determined by the
A0028801-EN
A0032216-EN
Page 56
Proline Promass F 100
purge system itself or by the device, depending on which component has the lower nominal pressure
classification.
If the device is fitted with a rupture disk (order code for "Sensor option", option CA "Rupture disk"), the
rupture disk trigger pressure is decisive for the maximum nominal pressure → 56.
The secondary containment burst pressure refers to a typical internal pressure achieved prior to
mechanical failure of the secondary containment as determined during type testing. The
corresponding type test declaration can be ordered with the device (order code for "Additional
approval", option LN "Type test containment").
DNSecondary containment
pressure rating
(designed with a safety factor
≥ 4)
[mm][in][bar][psi][bar][psi]
8³⁄₈405802553698
15½405802002900
251405802804060
401½405801802610
502405801952828
803253621051522
100416232851232
150616232801160
250101014557826
Secondary containment burst pressure
If a measuring tube fails (e.g. due to process characteristics like corrosive or abrasive fluids), the
fluid will be contained by the secondary containment.
If there is a need to drain the leaking medium into a discharge device, the sensor should be fitted
with a rupture disk. Connect the discharge to the additional threaded connection → 77.
If the sensor is to be purged with gas (gas detection), it should be equipped with purge connections.
Do not open the purge connections unless the containment can be filled immediately with a
dry, inert gas. Use only low pressure to purge. Maximum pressure: 5 bar (72.5 psi).
In case of a tube failure, the pressure level inside the secondary containment will rise according to
the operating process pressure. If the user judges that the secondary containment pressure rating/
burst pressure does not provide an adequate safety margin, the device can be fitted with a rupture
disk. This prevents excessively high pressure from forming inside the secondary containment.
Therefore, the use of a rupture disk is strongly recommended in applications involving high gas
pressures, and particularly in applications in which the process pressure is greater than 2/3 of the
secondary containment burst pressure.
For information on the dimensions: see the "Mechanical construction" section
Rupture disk
To increase the level of safety, a device version with a rupture disk with a trigger pressure of
10 to 15 bar (145 to 217.5 psi)can be used (order code for "Sensor option", option CA "rupture disk").
Rupture disks cannot be combined with the separately available heating jacket .
Special mounting instructions: → 47
For information on the dimensions: → 77
Flow limit
Select the nominal diameter by optimizing between the required flow range and permissible pressure
loss.
For an overview of the full scale values for the measuring range, see the "Measuring range"
section → 8
56Endress+Hauser
Page 57
Proline Promass F 100
• The minimum recommended full scale value is approx. 1/20 of the maximum full scale value
• In most applications, 20 to 50 % of the maximum full scale value can be considered ideal
• A low full scale value must be selected for abrasive media (such as liquids with entrained solids):
flow velocity < 1 m/s (< 3 ft/s).
• For gas measurement the following rules apply:
– The flow velocity in the measuring tubes should not exceed half the sound velocity (0.5 Mach).
– The maximum mass flow depends on the density of the gas: formula → 8
To calculate the flow limit, use the Applicator sizing tool → 100
Pressure loss
System pressure
Thermal insulation
To calculate the pressure loss, use the Applicator sizing tool → 100
Promass F with reduced pressure loss: order code for "Sensor option", option CE "reduced pressure
loss"
It is important that cavitation does not occur, or that gases entrained in the liquids do not outgas.
This is prevented by means of a sufficiently high system pressure.
For this reason, the following mounting locations are recommended:
• At the lowest point in a vertical pipe
• Downstream from pumps (no danger of vacuum)
A0028777
In the case of some fluids, it is important to keep the heat radiated from the sensor to the
transmitter to a low level. A wide range of materials can be used for the required insulation.
The following device versions are recommended for versions with thermal insulation:
• Version with extended neck for insulation:
Order code for "Sensor option", option CG with an extended neck length of 105 mm (4.13 in).
• Extended temperature version:
Order code for "Measuring tube material", option SD, SE, SF or TH with an extended neck length of
105 mm (4.13 in).
Heating
NOTICE
Electronics overheating on account of thermal insulation!
Maximum permissible temperature at the lower end of the transmitter housing: 80 °C (176 °F)
‣
Thermal insulation with extended neck free: the insulation is omitted around the extended neck.
‣
We recommend that you do not insulate the extended neck in order to ensure optimum
dissipation of heat.
37Thermal insulation with extended neck free
Some fluids require suitable measures to avoid loss of heat at the sensor.
A0034391
Endress+Hauser57
Page 58
Proline Promass F 100
Heating options
• Electrical heating, e.g. with electric band heaters
• Via pipes carrying hot water or steam
• Via heating jackets
Heating jackets for the sensors can be ordered as accessories from Endress+Hauser .→ 99
NOTICE
Danger of overheating when heating
Ensure that the temperature at the lower end of the transmitter housing does not exceed
‣
80 °C (176 °F).
Ensure that convection takes place on a sufficiently large scale at the transmitter neck.
‣
Ensure that a sufficiently large area of the housing support remains exposed. The uncovered part
‣
serves as a radiator and protects the electronics from overheating and excessive cooling.
Vibrations
The high oscillation frequency of the measuring tubes ensures that the correct operation of the
measuring system is not influenced by plant vibrations.
58Endress+Hauser
Page 59
Proline Promass F 100
A
B
C
G
D
L
E
F
K
M
Mechanical construction
Dimensions in SI unitsCompact version
A0033787
Order code for "Housing", option A "Compact, aluminum, coated"
DN
1)
[mm][mm][mm][mm][mm][mm][mm][mm][mm][mm][mm]
81489454751852601365.35
151489454751852601368.30
2514894547518526013612.0
401489454105189.5294.513617.6
501489454141199.5340.513626.0
801489454200219.5419.513640.5
100148945425423849213651.2
150148945437825963713668.9
2501489454548302.5850.5136102.3
1)Depending on the cable gland used: values up to + 30 mm
2)With order code for "Sensor option", option CG or order code for "Measuring tube material", option SD, SE,
SF, TH: values +70 mm
3)If using a display, order code for "Display; operation", option B: values +28 mm
4)Depending on respective process connection→ 61
1)
A
BCDE
2) 3)
F
2) 3)
GKLM
4)
4)
4)
4)
4)
4)
4)
4)
4)
70
70
70
79
99
139
176
218
305
Order code for "Housing", option B "Compact hygienic, stainless"
DN
1)
[mm][mm][mm][mm][mm][mm][mm][mm][mm][mm][mm]
81377859751802551345.35
151377859751802551348.30
2513778597518025513412.0
401377859105184.5289.513417.6
501377859141194.5335.513426.0
801377859200214.5414.513440.5
100137785925423348713451.2
1)
A
BCDE
2) 3)
F
2) 3)
GKLM
4)
4)
4)
4)
4)
4)
4)
70
70
70
79
99
139
176
Endress+Hauser59
Page 60
Proline Promass F 100
DN
1)
1)
A
BCDE
2) 3)
F
2) 3)
GKLM
[mm][mm][mm][mm][mm][mm][mm][mm][mm][mm][mm]
150137785937825463213468.9
2501377859548297.5845.5134102.3
4)
4)
218
305
1)Depending on the cable gland used: values up to + 30 mm
2)With order code for "Sensor option", option CG or order code for "Measuring tube material", option SD, SE,
SF, TH: values +70 mm
3)If using a display, order code for "Display; operation", option B: values +28 mm
4)Depending on respective process connection→ 61
Order code for "Housing", option C "Ultra-compact hygienic, stainless"
DN
1)
[mm][mm][mm][mm][mm][mm][mm][mm][mm][mm]
81246856751802551125.35
151246856751802551128.30
2512468567518025511212.0
401246856105184.5289.511217.6
501246856141194.5335.511226.0
801246856200214.5414.511240.5
100124685625423348711251.2
150124685637825463211268.9
2501246856548297.5845.5112102.3
1)
A
BCDF
2) 3)
F
2) 3)
GKLM
4)
4)
4)
4)
4)
4)
4)
4)
4)
139
176
218
305
70
70
70
79
99
1)Depending on the cable gland used: values up to + 30 mm
2)With order code for "Sensor option", option CG or order code for "Measuring tube material", option SD, SE,
SF, TH: values +70 mm
3)If using a display, order code for "Display; operation", option B: values +14 mm
4)Depending on respective process connection→ 61
60Endress+Hauser
Page 61
Proline Promass F 100
C
D
A
B
E
L
Flange connections
Fixed flange EN 1092-1, ASME B16.5, JIS B2220
A0015621
Length tolerance for dimension L in mm:
• DN ≤ 100: +1.5 / –2.0
• DN ≥ 125: +3.5
Flange according to EN 1092-1 (DIN 2501): PN16
1.4404 (F316/F316L): order code for "Process connection", option D1S
Alloy C22: order code for "Process connection", option D1C
Flange with groove according to EN 1092-1 Form D (DIN 2512N): PN16
1.4404 (F316/F316L): order code for "Process connection", option D5S
Alloy C22: order code for "Process connection", option D5C
DN
[mm]
1002201808 × ⌀1820107.11127/1400
1502852408 × ⌀2222159.31330/1700
25040535512 × ⌀2626260.41775
Surface roughness (flange): EN 1092-1 Form B1 (DIN 2526 Form C), Ra 3.2 to 12.5 µm
1)Installation length in accordance with NAMUR recommendation NE 132 optionally available (order code
for "Process connection", option D1N or D5N (with groove))
Flange according to EN 1092-1 (DIN 2501): PN16 with reduction in nominal diameter
1.4404 (F316/F316L
DN
[mm]
10080DHS2201808 × ⌀ 1820107.1874
150100DJS2852408 × ⌀ 2222159.31167
200150DLS34029512 × ⌀ 2224206.51461
Surface roughness (flange): EN 1092-1 Form B1 (DIN 2526 Form C), Ra 3.2 to 12.5 µm
A
[mm]
reduction
to DN
[mm]
B
[mm]
Order code for
"Process
connection",
option
C
[mm]
A
[mm]B[mm]
D
[mm]
C
[mm]
E
[mm]
D
[mm]E[mm]L[mm]
L
[mm]
1)
1)
Endress+Hauser61
Page 62
Proline Promass F 100
Flange according to EN 1092-1 (DIN 2501): PN 40
1.4404 (F316/F316L): order code for "Process connection", option D2S
Alloy C22: order code for "Process connection", option D2C
Flange with groove according to EN 1092-1 Form D (DIN 2512N): PN 40
1.4404 (F316/F316L): order code for "Process connection", option D6S
Alloy C22: order code for "Process connection", option D6C
DN
[mm]
1)
8
1595654 × ⌀141617.3404/510
25115854 × ⌀141828.5440/600
401501104 × ⌀181843.1550
501651254 × ⌀182054.5715
802001608 × ⌀182482.5840/915
1002351908 × ⌀2224107.11127
1503002508 × ⌀2628159.31370
25045038512 × ⌀3338258.81845
Surface roughness (flange): EN 1092-1 Form B1 (DIN 2526 Form C), Ra 3.2 to 12.5 µm
A
[mm]
B
[mm]
C
[mm]
D
[mm]
E
[mm]
95654 × ⌀141617.3370/510
L
[mm]
2)
2)
2)
2)
1)DN 8 with DN 15 flanges as standard
2)Installation length in accordance with NAMUR recommendation NE 132 optionally available (order code
for "Process connection", option D2N or D6N (with groove))
Flange according to EN 1092-1 (DIN 2501): PN 40 (with DN 25 flanges)
1.4404 (F316/F316L): order code for "Process connection", option R2S
DN
[mm]
A
[mm]
B
[mm]
C
[mm]
D
[mm]
E
[mm]
L
[mm]
8115854 × ⌀141828.5440
15115854 × ⌀141828.5440
Surface roughness (flange): EN 1092-1 Form B1 (DIN 2526 Form C), Ra 3.2 to 12.5 µm
Flange according to EN 1092-1 (DIN 2501): PN 40 with reduction in nominal diameter
1.4404 (F316/F316L)
DN
[mm]
reduction
to DN
[mm]
Order code for
"Process
connection",
A
[mm]B[mm]
C
[mm]
D
[mm]E[mm]L[mm]
option
5040DFS1651254 × ⌀ 182054.5555
8050DGS2001608 × ⌀ 182482.5840
10080DIS2351908 × ⌀ 2224107.1874
150100DKS3002508 × ⌀ 2628159.31167
200150DMS37532012 × ⌀ 3034206.51461
Surface roughness (flange): EN 1092-1 Form B1 (DIN 2526 Form C), Ra 3.2 to 12.5 µm
62Endress+Hauser
Page 63
Proline Promass F 100
Flange according to EN 1092-1 (DIN 2501): PN 63
1.4404 (F316/F316L): order code for "Process connection", option D3S
Alloy C22: order code for "Process connection", option D3C
Flange with groove according to EN 1092-1 Form D (DIN 2512N): PN 63
1.4404 (F316/F316L): order code for "Process connection", option D7S
Alloy C22: order code for "Process connection", option D7C
DN
[mm]
A
[mm]
B
[mm]
C
[mm]
D
[mm]
E
[mm]
[mm]
501801354 × ⌀222654.5724
802151708 × ⌀222881.7875
1002502008 × ⌀2630106.31127
1503452808 × ⌀3336157.11410
25047040012 × ⌀3646255.41885
Surface roughness (flange):
EN 1092-1 Form B1 (DIN 2526 Form C), Ra 3.2 to 12.5 µm EN 1092-1 Form B2 (DIN 2526 Form E), Ra
0.8 to 3.2 µm
Flange according to EN 1092-1 (DIN 2501): PN 100
1.4404 (F316/F316L): order code for "Process connection", option D4S
Alloy C22: order code for "Process connection", option D4C
Flange with groove according to EN 1092-1 Form D (DIN 2512N): PN 100
1.4404 (F316/F316L): order code for "Process connection", option D8S
Alloy C22: order code for "Process connection", option D8C
DN
[mm]
1)
8
A
[mm]
B
[mm]
C
[mm]
D
[mm]
E
[mm]
[mm]
105754 × ⌀142017.3400
15105754 × ⌀142017.3420
251401004 × ⌀182428.5470
401701254 × ⌀222642.5590
501951454 × ⌀262853.9740
802301808 × ⌀263280.9885
1002652108 × ⌀3036104.31127
15035529012 × ⌀3344154.01450
Surface roughness (flange): EN 1092-1 Form B2 (DIN 2526 Form E), Ra 0.8 to 3.2 µm
L
L
1)DN 8 with DN 15 flanges as standard
Flange according to ASME B16.5: Class 150
1.4404 (F316/F316L): order code for "Process connection", option AAS
Alloy C22: order code for "Process connection", option AAC
DN
[mm]
1)
8
A
[mm]
B
[mm]
C
[mm]
D
[mm]
E
[mm]
9060.34 × ⌀15.711.215.7370
L
[mm]
159060.34 × ⌀15.711.215.7404
2511079.44 × ⌀15.714.226.7440
4012598.44 × ⌀15.717.540.9550
50150120.74 × ⌀19.119.152.6715
80190152.44 × ⌀19.123.978.0840
100230190.58 × ⌀19.123.9102.41127
150280241.38 × ⌀22.425.4154.21398
Endress+Hauser63
Page 64
Proline Promass F 100
Flange according to ASME B16.5: Class 150
1.4404 (F316/F316L): order code for "Process connection", option AAS
Alloy C22: order code for "Process connection", option AAC
DN
[mm]
25040536212 × ⌀25.430.2254.51832
Surface roughness (flange): Ra 3.2 to 6.3 µm
1)DN 8 with DN 15 flanges as standard
Flange according to ASME B16.5: Class 150 with reduction in nominal diameter
1.4404 (F316/F316L)
DN
[mm]
5040AHS150120.74 × ⌀ 19.119.152.6550
8050AJS190152.44 × ⌀ 19.123.978.0720
10080ALS230190.58 × ⌀ 19.123.9102.4874
150100ANS280241.38 × ⌀ 22.425.4154.21167
200150APS345298.58 × ⌀ 22.429202.71461
Surface roughness (flange): Ra 3.2 to 6.3 µm
[mm]
reduction
to DN
[mm]
A
B
[mm]
Order code for
"Process
connection",
option
C
[mm]
A
[mm]B[mm]
D
[mm]
C
[mm]
E
[mm]
D
[mm]E[mm]L[mm]
L
[mm]
Flange according to ASME B16.5: Class 300
1.4404 (F316/F316L): order code for "Process connection", option ABS
Alloy C22: order code for "Process connection", option ABC
DN
[mm]
1)
8
159566.74 × ⌀15.714.215.7404
2512588.94 × ⌀19.117.526.7440
40155114.34 × ⌀22.320.640.9550
501651278 × ⌀19.122.352.6715
80210168.38 × ⌀22.328.478.0840
1002552008 × ⌀22.331.7102.41127
150320269.912 × ⌀22.336.5154.21417
250445387.416 × ⌀28.447.4254.51863
Surface roughness (flange): Ra 3.2 to 6.3 µm
1)DN 8 with DN 15 flanges as standard
Flange according to ASME B16.5: Class 300 with reduction in nominal diameter
1.4404 (F316/F316L)
DN
[mm]
5040AIS1651278 × ⌀ 19.122.352.6615
8050AKS210168.38 × ⌀ 22.328.478.0732
10080AMS2552008 × ⌀ 22.331.7102.4894
A
[mm]
9566.74 × ⌀15.714.215.7370
reduction
to DN
[mm]
B
[mm]
Order code for
"Process
connection",
option
C
[mm]
A
[mm]B[mm]
D
[mm]
C
[mm]
[mm]
D
[mm]E[mm]L[mm]
E
L
[mm]
64Endress+Hauser
Page 65
Proline Promass F 100
Flange according to ASME B16.5: Class 300 with reduction in nominal diameter
1.4404 (F316/F316L)
DN
[mm]
reduction
to DN
[mm]
Order code for
"Process
connection",
A
[mm]B[mm]
C
[mm]
D
[mm]E[mm]L[mm]
option
150100AOS320269.912 × ⌀ 22.336.5154.21187
200150AQS380330.212 × ⌀ 25.441.7202.71461
Surface roughness (flange): Ra 3.2 to 6.3 µm
Flange according to ASME B16.5: Class 600
1.4404 (F316/F316L): order code for "Process connection", option ACS
Alloy C22: order code for "Process connection", option ACC
DN
[mm]
1)
8
A
[mm]
B
[mm]
C
[mm]
D
[mm]
E
[mm]
9566.74 × ⌀15.720.613.9400
[mm]
159566.74 × ⌀15.720.613.9420
2512588.94 × ⌀19.123.924.3490
40155114.34 × ⌀22.328.738.1600
501651278 × ⌀19.131.849.2742
80210168.38 × ⌀22.338.273.7900
100275215.98 × ⌀25.448.497.31157
150355292.112 × ⌀28.447.8154.21467
250510431.816 × ⌀35.169.9254.51946
Surface roughness (flange): Ra 3.2 to 6.3 µm
L
1)DN 8 with DN 15 flanges as standard
Flange JIS B2220: 10K
1.4404 (F316/F316L): order code for "Process connection", option NDS
Alloy C22: order code for "Process connection", option NDC
DN
[mm]
A
[mm]
B
[mm]
C
[mm]
D
[mm]
E
[mm]
501551204 × ⌀191650715
801851508 × ⌀191880832
1002101758 × ⌀19181001127
1502802408 × ⌀23221501354
25040035512 × ⌀25242501775
Surface roughness (flange): Ra 3.2 to 6.3 µm
Flange JIS B2220: 20K
1.4404 (F316/F316L): order code for "Process connection", option NES
Alloy C22: order code for "Process connection", option NEC
DN
[mm]
1)
8
A
[mm]
B
[mm]
C
[mm]
D
[mm]
E
[mm]
95704 × ⌀151415370
1595704 × ⌀151415404
25125904 × ⌀191625440
401401054 × ⌀191840550
L
[mm]
L
[mm]
Endress+Hauser65
Page 66
Proline Promass F 100
Flange JIS B2220: 20K
1.4404 (F316/F316L): order code for "Process connection", option NES
Alloy C22: order code for "Process connection", option NEC
DN
[mm]
501551208 × ⌀191850715
802001608 × ⌀232280832
1002251858 × ⌀23241001127
15030526012 × ⌀25281501386
25043038012 × ⌀27342501845
Surface roughness (flange): Ra 1.6 to 3.2 µm
1)DN 8 with DN 15 flanges as standard
Flange JIS B2220: 40K
1.4404 (F316/F316L): order code for "Process connection", option NGS
Alloy C22: order code for "Process connection", option NGC
DN
[mm]
1)
8
15115804 × ⌀192015425
25130954 × ⌀192225485
401601204 × ⌀232438600
501651308 × ⌀192650760
802101708 × ⌀233275890
1002502058 × ⌀25361001167
15035529512 × ⌀33441501498
Surface roughness (flange): Ra 1.6 to 3.2 µm
A
[mm]
A
[mm]
B
[mm]
B
[mm]
C
[mm]
C
[mm]
D
[mm]
D
[mm]
E
[mm]
E
[mm]
115804 × ⌀192015400
L
[mm]
L
[mm]
1)DN 8 with DN 15 flanges as standard
Flange JIS B2220: 63K
1.4404 (F316/F316L): order code for "Process connection", option NHS
Alloy C22: order code for "Process connection", option NHC
DN
[mm]
1)
8
A
[mm]
B
[mm]
C
[mm]
D
[mm]
E
[mm]
120854 × ⌀192312420
15120854 × ⌀192312440
251401004 × ⌀232722494
401751304 × ⌀253235620
501851458 × ⌀233448775
802301858 × ⌀254073915
1002702208 × ⌀2744981167
15036530512 × ⌀33541461528
Surface roughness (flange): Ra 1.6 to 3.2 µm
1)DN 8 with DN 15 flanges as standard
L
[mm]
66Endress+Hauser
Page 67
Proline Promass F 100
C
D
A
B
E
XX
L
Fixed flange DIN 11864-2
A0015627
38Detail X: Asymmetrical process connection; the part shown in blue is provided by the supplier.
Length tolerance for dimension L in mm:
+1.5 / –2.0
Flange DIN11864-2 Form A, for pipe according to DIN11866 series A, flat with notch
1.4404 (316/316L)
Order code for "Process connection", option KCS
DN
[mm]
854374 × ⌀91010387
1559424 × ⌀91016418
2570534 × ⌀91026454
4082654 × ⌀91038560
5094774 × ⌀91050720
801331128 × ⌀111281900
1001591378 × ⌀11141001127
3A-version available: order code for "Additional approval", option LP in conjunction with
Ra ≤ 0.8 µm: order code for "Measuring tube material", option SB, SE or
Ra ≤ 0.4 µm: order code for "Measuring tube material", option SC, SF
A
[mm]
B
[mm]
C
[mm]
D
[mm]
E
[mm]
L
[mm]
Endress+Hauser67
Page 68
Proline Promass F 100
L
C
D
A
B
F
E
Lap joint flange EN 1092-1, ASME B16.5, JIS B2220
Length tolerance for dimension L in mm:
+1.5 / –2.0
Lap joint flange according to EN 1092-1 Form D: PN 40
1.4301 (F304), wetted parts Alloy C22Order code for "Process connection", option DAC
DN
[mm]A[mm]
2)
8
95654 × ⌀ 1414.54517.33700
1595654 × ⌀ 1414.54517.34040
25115854 × ⌀ 1416.56828.5444+4
401501104 × ⌀ 18218843.1560+10
501651254 × ⌀ 182310254.5719+4
802001608 × ⌀ 182913882.5848+8
1002351908 × ⌀ 2234162107.11131+4
Surface roughness (flange): Ra 3.2 to 12.5 µm
B
[mm]
C
[mm]
D
[mm]E[mm]
F
[mm]L[mm]
A0022221
L
diff
[mm]
1)
1)Difference to installation length of the welding neck flange (order code for "Process connection", option
D2C)
2)DN 8 with DN 15 flanges as standard
Lap joint flange according to ASME B16.5: Class 150
1.4301 (F304), wetted parts Alloy C22Order code for "Process connection", option ADC
DN
[mm]A[mm]
2)
8
9060.34 × ⌀ 15.71535.115.73700
B
[mm]
C
[mm]
D
[mm]E[mm]
F
[mm]L[mm]
L
diff
[mm]
1)
159060.34 × ⌀ 15.71535.115.74040
2511079.44 × ⌀ 15.71650.826.74400
4012598.44 × ⌀ 15.715.973.240.95500
50150120.74 × ⌀ 19.11991.952.67150
80190152.44 × ⌀ 19.122.3127.078.08400
68Endress+Hauser
Page 69
Proline Promass F 100
Lap joint flange according to ASME B16.5: Class 150
1.4301 (F304), wetted parts Alloy C22Order code for "Process connection", option ADC
DN
[mm]A[mm]
B
[mm]
C
[mm]
D
[mm]E[mm]F[mm]L[mm]
L
diff
[mm]
100230190.58 × ⌀ 19.126157.2102.411270
Surface roughness (flange): Ra 3.2 to 12.5 µm
1)Difference to installation length of the welding neck flange (order code for "Process connection", option
AAC)
2)DN 8 with DN 15 flanges as standard
Lap joint flange according to ASME B16.5: Class 300
1.4301 (F304), wetted parts Alloy C22Order code for "Process connection", option AEC
DN
[mm]A[mm]
2)
8
9566.74 × ⌀ 15.716.535.115.7376+6
B
[mm]
C
[mm]
D
[mm]E[mm]F[mm]L[mm]
L
diff
[mm]
159566.74 × ⌀ 15.716.535.115.7406+2
2512588.94 × ⌀ 19.121.050.826.7450+10
40155114.34 × ⌀ 22.323.073.240.9564+14
501651278 × ⌀ 19.125.591.952.6717+2
80210168.38 × ⌀ 22.331.0127.078.0852.6+12.6
1002552008 × ⌀ 22.332.0157.2102.41139+12
Surface roughness (flange): Ra 3.2 to 12.5 µm
1)
1)
1)Difference to installation length of the welding neck flange (order code for "Process connection", option
ABC)
2)DN 8 with DN 15 flanges as standard
Lap joint flange according to ASME B16.5: Class 600
1.4301 (F304), wetted parts Alloy C22Order code for "Process connection", option AFC
DN
[mm]A[mm]
2)
8
9566.74 × ⌀ 15.717.035.113.94000
B
[mm]
C
[mm]
D
[mm]E[mm]F[mm]L[mm]
L
diff
[mm]
159566.74 × ⌀ 15.717.035.113.94200
2512588.94 × ⌀ 19.121.550.824.34900
40155114.34 × ⌀ 22.325.073.238.16000
501651278 × ⌀ 19.128.091.949.27420
80210168.38 × ⌀ 22.335.0127.073.79000
100275215.98 × ⌀ 25.444.0157.297.31167+10
Surface roughness (flange): Ra 3.2 to 12.5 µm
1)Difference to installation length of the welding neck flange (order code for "Process connection", option
ACC)
2)DN 8 with DN 15 flanges as standard
1)
Endress+Hauser69
Page 70
Proline Promass F 100
Lap joint flange JIS B2220: 20K
1.4301 (F304), wetted parts Alloy C22Order code for "Process connection", option NIC
DN
[mm]
2)
8
1595704 × ⌀ 151451154040
25125904 × ⌀ 1918.567254400
401401054 × ⌀ 1918.581405500
501551208 × ⌀ 192396507150
802001608 × ⌀ 232913280844+12
1002251858 × ⌀ 232916010011270
Surface roughness (flange): Ra 3.2 to 12.5 µm
1)Difference to installation length of the welding neck flange (order code for "Process connection", option
NEC)
2)DN 8 with DN 15 flanges as standard
A
[mm]
B
[mm]
C
[mm]
D
[mm]
E
[mm]
F
[mm]
L
[mm]
95704 × ⌀ 151451153700
L
diff
[mm]
1)
70Endress+Hauser
Page 71
Proline Promass F 100
L
A
B
Clamp connections
Tri-Clamp
Length tolerance for dimension L in mm:
+1.5 / –2.0
Tri-Clamp (½"), for pipe according to DIN 11866 series C
1.4404 (316/316L)
Order code for "Process connection", option FDW
DN
[mm]
8½25.09.5367
15½25.09.5398
3-A version available: order code for "Additional approval", option LP in conjunction with
Ra ≤ 0.8 µm: order code for "Measuring tube material", option SB, SE or
Ra ≤ 0.4 µm: order code for "Measuring tube material", option SC, SF
Clamp
[in]
A
[mm]
B
[mm]
A0015625
L
[mm]
Tri-Clamp ( ≥ 1"), for pipe according to DIN 11866 series C
1.4404 (316/316L)
Order code for "Process connection", option FTS
DN
[mm]
8150.422.1367
15150.422.1398
25150.422.1434
401½50.434.8560
50263.947.5720
80390.972.9900
1004118.997.41127
3-A version available: order code for "Additional approval", option LP in conjunction with
Ra ≤ 0.8 µm: order code for "Measuring tube material", option SB, SE or
Ra ≤ 0.4 µm: order code for "Measuring tube material", option SC, SF
Clamp
[in]
A
[mm]
B
[mm]
L
[mm]
Endress+Hauser71
Page 72
Threaded couplings
L
A
B
Thread DIN 11851, DIN11864-1, SMS 1145
Length tolerance for dimension L in mm:
+1.5 / –2.0
Thread DIN 11851, for pipe according to DIN11866, series A
1.4404 (316/316L)
Order code for "Process connection", option FMW
DN
[mm]
8Rd 34 × ¹⁄₈16367
15Rd 34 × ¹⁄₈16398
25Rd 52 × ¹⁄₆26434
40Rd 65 × ¹⁄₆38560
50Rd 78 × ¹⁄₆50720
80Rd 110 × ¹⁄₄81900
100Rd 130 × ¹⁄₄1001127
3-A version available: order code for "Additional approval", option LP in conjunction with
Ra ≤ 0.8 µm: order code for "Measuring tube material", option SB, SE
A
[in]
B
[mm]
Proline Promass F 100
A0015628
L
[mm]
Thread DIN11864-1 Form A, for pipe according to DIN11866, series A
1.4404 (316/316L)
Order code for "Process connection", option FLW
DN
[mm]
8Rd 28 × ¹⁄₈10367
15Rd 34 × ¹⁄₈16398
25Rd 52 × ¹⁄₈26434
40Rd 65 × ¹⁄₆38560
50Rd 78 × ¹⁄₆50720
80Rd 110 × ¹⁄₄81900
100Rd 130 × ¹⁄₄1001127
3-A version available: order code for "Additional approval", option LP in conjunction with
Ra ≤ 0.8 µm: order code for "Measuring tube material", option SB, SE or
Ra ≤ 0.4 µm: order code for "Measuring tube material", option SC, SF
A
[in]
B
[mm]
L
[mm]
72Endress+Hauser
Page 73
Proline Promass F 100
Thread SMS 1145
1.4404 (316/316L)
Order code for "Process connection", option SCS
DN
[mm]
8Rd 40 × ¹⁄₆22.6367
15Rd 40 × ¹⁄₆22.6398
25Rd 40 × ¹⁄₆22.6434
40Rd 60 × ¹⁄₆35.6560
50Rd 70 × ¹⁄₆48.6720
80Rd 98 × ¹⁄₆72.9900
100Rd 132 × ¹⁄₆97.61127
3-A version available: order code for "Additional approval", option LP in conjunction with
Ra ≤ 0.8 µm: order code for "Measuring tube material", option SB, SE
A
[in]
B
[mm]
L
[mm]
Endress+Hauser73
Page 74
Thread ISO 2853
L
A
B
Length tolerance for dimension L in mm:
+1.5 / –2.0
Thread ISO 2853, for pipe according to ISO 2037
1.4404 (316/316L)
Order code for "Process connection", option JSF
DN
[mm]
837.1322.6367
1537.1322.6398
2537.1322.6434
4052.6835.6560
5064.1648.6720
8091.1972.9900
100118.2197.61127
3-A version available: order code for "Additional approval", option LP in conjunction with
Ra ≤ 0.8 µm: order code for "Measuring tube material", option SB, SE or
Ra ≤ 0.4 µm: order code for "Measuring tube material", option SC, SF
1)
A
[mm]
B
[mm]
Proline Promass F 100
A0015623
L
[mm]
1)Max. thread diameter as per ISO 2853 annex A
74Endress+Hauser
Page 75
Proline Promass F 100
L
A
B
VCO
Length tolerance for dimension L in mm:
+1.5 / –2.0
8-VCO-4 (½")
1.4404 (316/316L)
Order code for "Process connection", option CVS
DN
[mm]
8AF 110.2390
A
[in]
B
[mm]
A0015624
L
[mm]
12-VCO-4 (¾")
1.4404 (316/316L)
Order code for "Process connection", option CWS
DN
[mm]
15AF 1½15.7430
A
[in]
B
[mm]
L
[mm]
Endress+Hauser75
Page 76
Safety Barrier Promass 100
B
D
A
C
Top-hat rail EN 60715:
• TH 35 x 7.5
• TH 35 x 15
ABCD
[mm][mm][mm][mm]
108114.59922.5
Proline Promass F 100
A0016777
76Endress+Hauser
Page 77
Proline Promass F 100
LL
45°
A
A
SW 1"
2
½" NPT
1
½" NPT
35 (1.38)ca. 75 (approx. 2.95)
1
1;2
1
1;2
Accessories
Rupture disk/purge connections
39
1Connection nipple for purge connections: order code for "Sensor options", option CH "Purge connection"
2Connection nipple with rupture disk: order code for "Sensor option", option CA "Rupture disk"
DNAL
[mm][mm][mm]
862216
1562220
2562260
4067310
5079452
80101560
100120684
150141880
250182380
A0028914
Endress+Hauser77
Page 78
Dimensions in US unitsCompact version
A
B
C
G
D
L
E
F
K
M
Order code for "Housing", option A "Compact, aluminum, coated"
DNA
[in][in][in][in][in][in][in][in][in][in][in]
³⁄₈5.833.702.132.957.2810.245.350.211
½5.833.702.132.957.2810.245.350.33
15.833.702.132.957.2810.245.350.47
1½5.833.702.134.137.4611.595.350.69
25.833.702.135.557.8513.415.351.02
35.833.702.137.878.6416.525.351.59
45.833.702.13109.3719.375.352.02
65.833.702.1314.8810.225.085.352.71
105.833.702.1321.5711.9133.485.354.03
1)
Proline Promass F 100
A0033787
1)
B
CDF
2) 3)
F
2) 3)
GKLM
4)
4)
4)
4)
4)
4)
4)
4)
4)
2.76
2.76
2.76
3.11
3.90
5.47
6.93
8.58
12.01
1)Depending on the cable gland used: values up to + 1.18 in
2)With order code for "Sensor option", option CG or order code for "Measuring tube material", option SD, SE,
SF, TH: values +2.76 in
3)If using a display, order code for "Display; operation", option B: values +1.1 in
4)Depending on respective process connection→ 80
Order code for "Housing", option B "Compact hygienic, stainless"
DNA
1)
BCDF
[in][in][in][in][in][in][in][in][in][in][in]
³⁄₈5.393.072.322.957.0910.045.280.211
½5.393.072.322.957.0910.045.280.33
15.393.072.322.957.0910.045.280.47
1½5.393.072.324.137.2611.45.280.69
25.393.072.325.557.6613.215.281.02
35.393.072.327.878.4416.325.281.59
45.393.072.32109.1719.175.282.02
2) 3)
F
2) 3)
GKLM
4)
4)
4)
4)
4)
4)
4)
2.76
2.76
2.76
3.11
3.90
5.47
6.93
78Endress+Hauser
Page 79
Proline Promass F 100
DNA
1)
BCDF
2) 3)
F
2) 3)
GKLM
[in][in][in][in][in][in][in][in][in][in][in]
65.393.072.3214.881024.885.282.71
105.393.072.3221.5711.7133.295.284.03
4)
4)
8.58
12.01
1)Depending on the cable gland used: values up to +1.18 in
2)With order code for "Sensor option", option CG or order code for "Measuring tube material", option SD, SE,
SF, TH: values +2.76 in
3)If using a display, order code for "Display; operation", option B: values +1.1 in
4)Depending on respective process connection→ 80
Order code for "Housing", option C "Ultra-compact hygienic, stainless"
DNA
1)
BCDF
[in][in][in][in][in][in][in][in][in][in][in]
³⁄₈4.882.682.202.957.0910.044.410.21
½4.882.682.202.957.0910.044.410.33
14.882.682.202.957.0910.044.410.47
1½4.882.682.204.137.2611.44.410.69
24.882.682.205.557.6613.214.411.02
34.882.682.207.878.4416.324.411.59
44.882.682.20109.1719.174.412.02
64.882.682.2014.881024.884.412.71
104.882.682.2021.5711.7133.294.414.03
2) 3)
F
2) 3)
GKLM
4)
4)
4)
4)
4)
4)
4)
4)
4)
2.76
2.76
2.76
3.11
3.90
5.47
6.93
8.58
12.01
1)Depending on the cable gland used: values up to +1.18 in
2)With order code for "Sensor option", option CG or order code for "Measuring tube material", option SD, SE,
SF, TH: values +2.76 in
3)If using a display, order code for "Display; operation", option B: values +1.1 in
4)Depending on respective process connection→ 80
Endress+Hauser79
Page 80
Flange connections
C
D
A
B
E
L
Fixed flange ASME B16.5
Length tolerance for dimension L in inch:
• DN ≤ 4": +0.06 / –0.08
• DN ≥ 5": +0.14
Proline Promass F 100
A0015621
Flange according to ASME B16.5: Class 150
1.4404 (F316/F316L): order code for "Process connection", option AAS
Alloy C22: order code for "Process connection", option AAC
DN
[in]
1)
³⁄₈
½3.542.374 × ⌀0.620.440.6215.91
14.333.134 × ⌀0.620.561.0517.32
1½4.923.874 × ⌀0.620.691.6121.65
25.914.754 × ⌀0.750.752.0728.15
37.486.004 × ⌀0.750.943.0733.07
49.067.508 × ⌀0.750.944.0344.37
611.029.508 × ⌀0.8816.0755.04
1015.9414.2512 × ⌀1.01.1910.0272.13
Surface roughness (flange): Ra 125 to 250 µin
1)DN ³⁄₈" with DN ½" flanges as standard
Flange according to ASME B16.5: Class 150 with reduction in nominal diameter
1.4404 (F316/F316L)
DN
[in]
21½AHS5.914.754 × ⌀ 0.750.752.0721.65
32AJS7.4864 × ⌀ 0.750.943.0728.35
43ALS9.067.58 × ⌀ 0.750.944.0334.41
64ANS11.029.58 × ⌀ 0.8816.0745.94
86APS13.5811.758 × ⌀ 0.881.147.9857.52
Surface roughness (flange): Ra 125 to 250 µin
A
[in]
3.542.374 × ⌀ 0.620.440.6214.57
reduction
to DN
[in]
Order code for
connection",
[in]
"Process
option
B
A
[in]
C
[in]
B
[in]
D
[in]
C
[in]
E
[in]
D
[in]
E
[in]
L
[in]
L
[in]
80Endress+Hauser
Page 81
Proline Promass F 100
Flange according to ASME B16.5: Class 300
1.4404 (F316/F316L): order code for "Process connection", option ABS
Alloy C22: order code for "Process connection", option ABC
DN
[in]
³⁄₈
1)
A
[in]
B
[in]
C
[in]
D
[in]
E
[in]
3.742.634 × ⌀0.620.560.6214.57
½3.742.634 × ⌀0.620.560.6215.91
14.923.504 × ⌀0.750.691.0517.32
1½6.104.504 × ⌀0.880.811.6121.65
26.505.008 × ⌀0.750.882.0728.15
38.276.638 × ⌀0.881.123.0733.07
410.047.878 × ⌀0.881.254.0344.37
612.610.6312 × ⌀0.881.446.0755.79
1017.5215.2516 × ⌀1.121.8710.0273.35
Surface roughness (flange): Ra 125 to 250 µin
1)DN ³⁄₈" with DN ½" flanges as standard
L
[in]
Flange according to ASME B16.5: Class 300 with reduction in nominal diameter
1.4404 (F316/F316L)
DN
[in]
reduction
to DN
[in]
Order code for
"Process
connection",
A
[in]
B
[in]
C
[in]
D
[in]
E
[in]
option
21½AIS6.558 × ⌀ 0.750.882.0724.21
32AKS8.276.638 × ⌀ 0.881.123.0728.82
43AMS10.047.878 × ⌀ 0.881.254.0335.2
64AOS12.610.6312 × ⌀ 0.881.446.0746.73
86AQS14.961312 × ⌀ 11.647.9857.52
Surface roughness (flange): Ra 125 to 250 µin
Flange according to ASME B16.5: Class 600
1.4404 (F316/F316L): order code for "Process connection", option ACS
Alloy C22: order code for "Process connection", option ACC
DN
[in]
³⁄₈
1)
A
[in]
B
[in]
C
[in]
D
[in]
E
[in]
3.742.634 × ⌀0.620.810.5515.75
½3.742.634 × ⌀0.620.810.5516.54
14.923.504 × ⌀0.750.940.9619.29
1½6.104.504 × ⌀0.881.131.523.62
26.505.008 × ⌀0.751.251.9429.21
38.276.638 × ⌀0.881.52.935.43
410.838.508 × ⌀1.001.913.8345.55
613.9811.5012 × ⌀1.121.886.0757.76
1020.0817.0016 × ⌀1.382.7510.0276.61
Surface roughness (flange): Ra 125 to 250 µin
L
[in]
L
[in]
1)DN ³⁄₈" with DN ½" flanges as standard
Endress+Hauser81
Page 82
Proline Promass F 100
L
C
D
A
B
F
E
Lap joint flange ASME B16.5
Length tolerance for dimension L in inch:
+0.06 / –0.08
Lap joint flange according to ASME B16.5: Class 150
1.4301 (F304), wetted parts Alloy C22Order code for "Process connection", option ADC
DN
[in]
2)
³⁄₈
½3.542.374 × ⌀ 0.620.591.380.6215.910
14.333.134 × ⌀ 0.620.6321.0517.320
1½4.923.874 × ⌀ 0.620.632.881.6121.650
25.914.754 × ⌀ 0.750.753.622.0728.150
37.486.004 × ⌀ 0.750.8853.0733.070
49.067.508 × ⌀ 0.751.026.194.0344.370
Surface roughness (flange): Ra 125 to 492 µin
A
[in]
B
[in]
C
[in]
D
[in]
E
[in]
F
[in]
L
[in]
3.542.374 × ⌀ 0.620.591.380.6214.570
L
diff
[in]
A0022221
1)
1)Difference to installation length of the welding neck flange (order code for "Process connection", option
AAC)
2)DN ³⁄₈" with DN ½" flanges as standard
Lap joint flange according to ASME B16.5: Class 300
1.4301 (F304), wetted parts Alloy C22Order code for "Process connection", option AEC
DN
[in]
³⁄₈
A
[in]
2)
3.742.634 × ⌀ 0.620.651.380.6214.8+0.23
B
[in]
C
[in]
D
[in]
E
[in]
F
[in]
L
[in]
L
diff
[in]
1)
½3.742.634 × ⌀ 0.620.651.380.6215.98+0.07
14.923.504 × ⌀ 0.750.8321.0517.72+0.40
1½6.104.504 × ⌀ 0.880.912.881.6122.2+0.55
26.505.008 × ⌀ 0.7513.622.0728.23+0.08
38.276.638 × ⌀ 0.881.2253.0733.57+0.50
82Endress+Hauser
Page 83
Proline Promass F 100
Lap joint flange according to ASME B16.5: Class 300
1.4301 (F304), wetted parts Alloy C22Order code for "Process connection", option AEC
DN
[in]
A
[in]
B
[in]
C
[in]
D
[in]
E
[in]
F
[in]
L
[in]
L
diff
[in]
410.047.878 × ⌀ 0.881.266.194.0344.84+0.47
Surface roughness (flange): Ra 125 to 492 µin
1)Difference to installation length of the welding neck flange (order code for "Process connection", option
AAC)
2)DN ³⁄₈" with DN ½" flanges as standard
Lap joint flange according to ASME B16.5, Class 600
1.4301 (F304), wetted parts Alloy C22Order code for "Process connection", option AFC
DN
[in]
³⁄₈
2)
A
[in]
B
[in]
C
[in]
D
[in]
E
[in]
F
[in]
L
[in]
L
diff
[in]
3.742.634 × ⌀ 0.620.671.380.5515.750
½3.742.634 × ⌀ 0.620.671.380.5516.540
14.923.504 × ⌀ 0.750.8520.9619.290
1½6.104.504 × ⌀ 0.880.982.881.523.620
26.505.008 × ⌀ 0.751.13.621.9429.210
38.276.638 × ⌀ 0.881.3852.935.430
410.838.508 × ⌀ 11.736.193.8345.94+0.39
Surface roughness (flange): Ra 125 to 492 µin
1)
1)
1)Difference to installation length of the welding neck flange (order code for "Process connection", option
AAC)
2)DN ³⁄₈" with DN ½" flanges as standard
Endress+Hauser83
Page 84
Clamp connections
L
A
B
Tri-Clamp
Length tolerance for dimension L in inch:
+0.06 / –0.08
Tri-Clamp (½"), DIN 11866 series C
1.4404 (316/316L)
Order code for "Process connection", option FDW
DN
[in]
³⁄₈½0.980.3714.4
½½0.980.3715.7
3-A version available: order code for "Additional approval", option LP in conjunction with
Ra ≤ 32 µin: order code for "Measuring tube material", option SB, SE or
Ra ≤ 16 µin: order code for "Measuring tube material", option SC, SF
Clamp
[in]
A
[in]
B
[in]
Proline Promass F 100
A0015625
L
[in]
Tri-Clamp ( ≥ 1"), DIN 11866 series C
1.4404 (316/316L)
Order code for "Process connection", option FTS
DN
[in]
³⁄₈11.980.8714.4
½11.980.8715.7
111.980.8717.1
1½1½1.981.3722.0
222.521.8728.3
333.582.8735.4
444.683.8344.4
3-A version available: order code for "Additional approval", option LP in conjunction with
Ra ≤ 32 µin: order code for "Measuring tube material", option SB, SE or
Ra ≤ 16 µin: order code for "Measuring tube material", option SC, SF
Clamp
[in]
A
[in]
B
[in]
L
[in]
84Endress+Hauser
Page 85
Proline Promass F 100
L
A
B
Threaded couplings
Thread SMS 1145
Length tolerance for dimension L in inch:
+0.06 / –0.08
Thread SMS 1145
1.4404 (316/316L)
Order code for "Process connection", option SCS
DN
[in]
³⁄₈Rd 40 × ¹⁄₆0.8914.45
½Rd 40 × ¹⁄₆0.8915.67
1Rd 40 × ¹⁄₆0.8917.09
1½Rd 60 × ¹⁄₆1.422.05
2Rd 70 × ¹⁄₆1.9128.35
3Rd 98 × ¹⁄₆2.8735.43
4Rd 132 × ¹⁄₆3.8444.37
3-A version available: order code for "Additional approval", option LP in conjunction with
Ra ≤ 32 µin: order code for "Measuring tube material", option SB, SE
A
[in]
B
[in]
A0015628
L
[in]
Endress+Hauser85
Page 86
VCO
L
A
B
B
D
A
C
Length tolerance for dimension L in inch:
+0.06 / –0.08
8-VCO-4 (½")
1.4404 (316/316L)
Order code for "Process connection", option CVS
DN
[in]
³⁄₈AF 10.415.35
A
[in]
B
[in]
Proline Promass F 100
A0015624
L
[in]
12-VCO-4 (¾")
1.4404 (316/316L)
Order code for "Process connection", option CWS
DN
[in]
½AF 1½0.6216.93
A
[in]
Safety Barrier Promass 100
Top-hat rail EN 60715:
• TH 35 x 7.5
• TH 35 x 15
B
[in]
L
[in]
ABCD
[in][in][in][in]
4.254.513.90.89
86Endress+Hauser
A0016777
Page 87
Proline Promass F 100
LL
45°
A
A
SW 1"
2
½" NPT
1
½" NPT
35 (1.38)ca. 75 (approx. 2.95)
1
1;2
1
1;2
Accessories
Rupture disk/purge connections
Weight
A0028914
1Connection nipple for purge connections: order code for "Sensor options", option CH "Purge connection"
2Connection nipple with rupture disk: order code for "Sensor option", option CA "Rupture disk"
DNAL
[in][in][in]
³⁄₈2.448.50
½2.448.66
12.4410.24
1½2.6412.20
23.1117.78
33.9822.0
44.7227.0
65.5534.6
107.1714.96
All values (weight exclusive of packaging material) refer to devices with EN/DIN PN 40 flanges.
Weight specifications including transmitter: order code for "Housing", option A "Compact, aluminum
coated".
Different values due to different transmitter versions:
Endress+Hauser87
Weight in SI units
DN [mm]Weight [kg]
89
1510
2512
4017
5028
8053
10094
Page 88
DN [mm]Weight [kg]
1
2
3
Weight in US units
DN [in]Weight [lbs]
Proline Promass F 100
150152
250398
3/820
½22
126
1½37
262
3117
4207
6335
10878
Safety Barrier Promass 100
49 g (1.73 ounce)
MaterialsTransmitter housing
• Order code for "Housing", option A "Compact, aluminum coated":
Aluminum, AlSi10Mg, coated
• Order code for "Housing", option B "Compact, hygienic, stainless":
– Hygienic version, stainless steel 1.4301 (304)
– Optional: order code for "Sensor feature", option CC
Hygienic version, for maximum corrosion resistance: stainless steel 1.4404 (316L)
• Order code for "Housing", option C "Ultra-compact, hygienic, stainless":
– Hygienic version, stainless steel 1.4301 (304)
– Optional: order code for "Sensor feature", option CC
Hygienic version, for maximum corrosion resistance: stainless steel 1.4404 (316L)
• Window material for optional local display (→ 91):
– For order code for "Housing", option A: glass
– For order code for "Housing", option B and C: plastic
Cable entries/cable glands
40Possible cable entries/cable glands
1Female thread M20 × 1.5
2Cable gland M20 × 1.5
3Adapter for cable entry with internal thread G ½" or NPT ½"
88Endress+Hauser
A0020640
Page 89
Proline Promass F 100
Order code for "Housing", option A "Compact, aluminum, coated"
The various cable entries are suitable for hazardous and non-hazardous areas.
Cable entry/cable glandMaterial
Cable gland M20 × 1.5
Nickel-plated brassAdapter for cable entry with internal thread G ½"
Adapter for cable entry with internal thread NPT ½"
Order code for "Housing", option B "Compact hygienic, stainless"
The various cable entries are suitable for hazardous and non-hazardous areas.
• DN 8 to 250 (3/8 to 10"): Alloy C22, 2.4602 (UNS N06022);
Manifold: Alloy C22, 2.4602 (UNS N06022)
Endress+Hauser89
Page 90
Proline Promass F 100
Process connections
• Flanges according to EN 1092-1 (DIN2501) / according to ASME B 16.5 / as per JIS B2220:
– Stainless steel, 1.4404 (F316/F316L)
– Alloy C22, 2.4602 (UNS N06022)
– Lap joint flanges: stainless steel, 1.4301 (F304); wetted parts Alloy C22
• All other process connections:
Stainless steel, 1.4404 (316/316L)
Available process connections→ 90
Seals
Welded process connections without internal seals
Safety Barrier Promass 100
Housing: Polyamide
Process connections
Surface roughness
• Fixed flange connections:
– EN 1092-1 (DIN 2501) flange
– EN 1092-1 (DIN 2512N) flange
– Namur lengths in accordance with NE 132
– ASME B16.5 flange
– JIS B2220 flange
– DIN 11864-2 Form A flange, DIN 11866 series A, flange with notch
• Clamp connections:
Tri-Clamp (OD tubes), DIN 11866 series C
• Thread:
– DIN 11851 thread, DIN 11866 series A
– SMS 1145 thread
– ISO 2853 thread, ISO 2037
– DIN 11864-1 Form A thread, DIN 11866 series A
• VCO connections:
– 8-VCO-4
– 12-VCO-4
Process connection materials
All data relate to parts in contact with fluid. The following surface roughness quality can be ordered.
• Not polished
• Ra
= 0.8 µm (32 µin)
max
• Ra
= 0.4 µm (16 µin)
max
Operability
Operating concept
90Endress+Hauser
Operator-oriented menu structure for user-specific tasks
• Commissioning
• Operation
• Diagnostics
• Expert level
Quick and safe commissioning
• Individual menus for applications
• Menu guidance with brief explanations of the individual parameter functions
Page 91
Proline Promass F 100
1235
7
46
Reliable operation
• Operation in the following languages:
– Via "FieldCare", "DeviceCare" operating tool:
English, German, French, Spanish, Italian, Chinese, Japanese
– Via integrated Web browser(only available for device versions with HART, PROFIBUS DP,
PROFINET and EtherNet/IP):
English, German, French, Spanish, Italian, Dutch, Portuguese, Polish, Russian, Turkish, Chinese,
Japanese, Bahasa (Indonesian), Vietnamese, Czech, Swedish, Korean
• Uniform operating philosophy applied to operating tools and Web browser
• If replacing the electronic module, transfer the device configuration via the plug-in memory
(HistoROM DAT) which contains the process and measuring device data and the event logbook. No
need to reconfigure.
For devices with Modbus RS485, the data recovery function is implemented without the plug-in
memory (HistoROM DAT).
• Troubleshooting measures can be called up via the operating tools and web browser
• Diverse simulation options
• Status indicated by several light emitting diodes (LEDs) on the electronic module in the housing
compartment
Local display
A local display is only available for device versions with the following communication protocols:
HART, PROFIBUS-DP, PROFINET, EtherNet/IP
The local display is only available with the following device order code:
Order code for "Display; operation", option B: 4-line; illuminated, via communication
Display element
• 4-line liquid crystal display with 16 characters per line.
• White background lighting; switches to red in event of device errors.
• Format for displaying measured variables and status variables can be individually configured.
• Permitted ambient temperature for the display: –20 to +60 °C (–4 to +140 °F). The readability of
the display may be impaired at temperatures outside the temperature range.
Remote operationVia HART protocol
This communication interface is available in device versions with a HART output.
41Options for remote operation via HART protocol
1Control system (e.g. PLC)
2Field Communicator 475
3Computer with operating tool (e.g. FieldCare, AMS Device Manager, SIMATIC PDM)
4Commubox FXA195 (USB)
5Field Xpert SFX350 or SFX370
6VIATOR Bluetooth modem with connecting cable
7Transmitter
Via PROFIBUS DP network
This communication interface is available in device versions with PROFIBUS DP.
Endress+Hauser91
A0028747
Page 92
12
3
444
42Options for remote operation via PROFIBUS DP network
4
213
555
1Automation system
2Computer with PROFIBUS network card
3PROFIBUS DP network
4Measuring device
Proline Promass F 100
A0020903
Via EtherNet/IP network
This communication interface is available in device versions with EtherNet/IP.
Star topology
43Options for remote operation via EtherNet/IP network: star topology
1Automation system, e.g. "RSLogix" (Rockwell Automation)
2Workstation for measuring device operation: with Custom Add-On Profile for "RSLogix 5000" (Rockwell
Automation) or with Electronic Data Sheet (EDS)
3Computer with Web browser (e.g. Internet Explorer) for accessing the integrated device Web server or
computer with operating tool (e.g. FieldCare, DeviceCare) with COM DTM "CDI Communication TCP/IP"
4Ethernet switch
5Measuring device
A0032078
Via PROFINET network
This communication interface is available in device versions with PROFINET.
92Endress+Hauser
Page 93
Proline Promass F 100
312
444
1
Star topology
A0026545
44Options for remote operation via PROFINET network: star topology
1Automation system, e.g. Simatic S7 (Siemens)
2Computer with Web browser (e.g. Internet Explorer) for accessing the integrated device Web server or
computer with operating tool (e.g. FieldCare, DeviceCare, SIMATIC PDM) with COM DTM "CDI Communication
TCP/IP"
3Switch, e.g. Scalance X204 (Siemens)
4Measuring device
Service interfaceVia service interface (CDI-RJ45)
This communication interface is present in the following device version:
• Order code for "Output", option B: 4-20 mA HART, pulse/frequency/switch output
• Order code for "Output", option L: PROFIBUS DP
• Order code for "Output", option N: EtherNet/IP
• Order code for "Output", option R: PROFINET
HART
45Connection for the order code for "Output", option B: 4-20 mA HART, pulse/frequency/switch output
1Service interface (CDI -RJ45) of the measuring device with access to the integrated Web server
2Computer with Web browser (e.g. Internet Explorer) for accessing the integrated device Web server or with
"FieldCare" operating tool with COM DTM "CDI Communication TCP/IP"
3Standard Ethernet connecting cable with RJ45 plug
A0016926
Endress+Hauser93
Page 94
Proline Promass F 100
1
1
PROFIBUS DP
A0021270
46Connection for order code for "Output", option L: PROFIBUS DP
1Service interface (CDI -RJ45) of the measuring device with access to the integrated Web server
2Computer with Web browser (e.g. Internet Explorer) for accessing the integrated device Web server or with
"FieldCare" operating tool with COM DTM "CDI Communication TCP/IP"
3Standard Ethernet connecting cable with RJ45 plug
EtherNet/IP
A0016940
47Connection for order code for "Output", option N: EtherNet/IP
1Service interface (CDI -RJ45) and EtherNet/IP interface of the measuring device with access to the integrated
Web server
2Computer with Web browser (e.g. Internet Explorer) for accessing the integrated device Web server or with
"FieldCare" operating tool with COM DTM "CDI Communication TCP/IP"
3Standard Ethernet connecting cable with RJ45 plug
94Endress+Hauser
Page 95
Proline Promass F 100
1
1
2
PROFINET
A0016940
48Connection for order code for "Output", option R: PROFINET
1Service interface (CDI -RJ45) and PROFINET interface of the measuring device with access to the integrated
Web server
2Computer with Web browser (e.g. Internet Explorer) for accessing the integrated device Web server or with
"FieldCare" operating tool with COM DTM "CDI Communication TCP/IP"
3Standard Ethernet connecting cable with RJ45 plug
Via service interface (CDI)
This communication interface is present in the following device version:
Order code for "Output", option M: Modbus RS485
Modbus RS485
1Service interface (CDI) of measuring device
2Commubox FXA291
3Computer with "FieldCare" operating tool with COM DTM "CDI Communication FXA291"
A0030216
Certificates and approvals
CE mark
C-Tick symbol
Endress+Hauser95
The device meets the legal requirements of the applicable EU Directives. These are listed in the
corresponding EU Declaration of Conformity along with the standards applied.
Endress+Hauser confirms successful testing of the device by affixing to it the CE mark.
The measuring system meets the EMC requirements of the "Australian Communications and Media
Authority (ACMA)".
Page 96
Proline Promass F 100
Ex approval
The measuring device is certified for use in hazardous areas and the relevant safety instructions are
provided in the separate "Safety Instructions" (XA) document. Reference is made to this document on
the nameplate.
The separate Ex documentation (XA) containing all the relevant explosion protection data is
available from your Endress+Hauser sales center.
ATEX/IECEx
Currently, the following versions for use in hazardous areas are available:
Ex ia
Category (ATEX)Type of protection
II1/2GEx ia IIC T6…T1 Ga/Gb or Ex ia IIB T6…T1 Ga/Gb
II2GEx ia IIC T6…T1 Gb or Ex ia IIB T6…T1 Gb
II1/2G, II2DEx ia IIC T6…T1 Ga/Gb or Ex ia IIB T6…T1 Ga/Gb
Ex tb IIIC Txx °C Db
II2G, II2DEx ia IIC T6…T1 Gb or Ex ia IIB T6…T1 Gb
Ex tb IIIC Txx °C Db
Ex nA
Category (ATEX)Type of protection
II3GEx nA IIC T6…T1 Gc or Ex nA IIC T5-T1 Gc
CSA
C
US
Currently, the following versions for use in hazardous areas are available:
IS (Ex i)
• Class I Division 1 Groups ABCD
• Class II Division 1 Groups EFG and Class III
NI (Ex nA)
Class I Division 2 Groups ABCD
Sanitary compatibility
• 3-A approval
Only devices with the order code for "Additional approval", option LP "3A" have 3-A approval.
• EHEDG-tested
Only devices with the order code for "Additional approval", option LT "EHEDG" have been tested and
meet the requirements of the EHEDG.
To meet the requirements for EHEDG certification, the device must be used with process
connections in accordance with the EHEDG position paper entitled "Easy Cleanable Pipe Couplings
and Process Connections" (www.ehedg.org).
HART certificationHART interface
The measuring device is certified and registered by the FieldComm Group. The measuring system
meets all the requirements of the following specifications:
• Certified according to HART 7
• The device can also be operated with certified devices of other manufacturers (interoperability)
Certification PROFIBUSPROFIBUS interface
The measuring device is certified and registered by the PNO (PROFIBUS User Organization
Organization). The measuring system meets all the requirements of the following specifications:
• Certified in accordance with PROFIBUS PA Profile 3.02
• The device can also be operated with certified devices of other manufacturers (interoperability)
96Endress+Hauser
Page 97
Proline Promass F 100
Certification PROFINETPROFINET interface
The measuring device is certified and registered by the PNO (PROFIBUS User Organization
Organization). The measuring system meets all the requirements of the following specifications:
• Certified according to:
– Test specification for PROFINET devices
– PROFINET Security Level 1 – Netload Class
• The device can also be operated with certified devices of other manufacturers (interoperability)
EtherNet/IP certification
Modbus RS485 certification
Pressure Equipment
Directive
The measuring device is certified and registered by the ODVA (Open Device Vendor Association).
The measuring system meets all the requirements of the following specifications:
• Certified in accordance with the ODVA Conformance Test
• EtherNet/IP Performance Test
• EtherNet/IP PlugFest compliance
• The device can also be operated with certified devices of other manufacturers (interoperability)
The measuring device meets all the requirements of the MODBUS/TCP conformity test and has the
"MODBUS/TCP Conformance Test Policy, Version 2.0". The measuring device has successfully passed
all the test procedures carried out.
The devices can be ordered with or without a PED approval. If a device with a PED approval is
required, this must be explicitly stated in the order. For devices with nominal diameters less than or
equal to DN 25 (1"), this is neither possible nor necessary.
• With the identification PED/G1/x (x = category) on the sensor nameplate, Endress+Hauser
confirms conformity with the "Essential Safety Requirements" specified in Appendix I of the
Pressure Equipment Directive 2014/68/EU.
• Devices bearing this marking (PED) are suitable for the following types of medium:
– Media in Group 1 and 2 with a vapor pressure greater than, or smaller and equal
to0.5 bar (7.3 psi)
– Unstable gases
• Devices not bearing this marking (PED) are designed and manufactured according to good
engineering practice. They meet the requirements of Article 4 paragraph 3 of the Pressure
Equipment Directive 2014/68/EU. The range of application is indicated in tables 6 to 9 in Annex II
of the Pressure Equipment Directive 2014/68/EU.
Other standards and
guidelines
• EN 60529
Degrees of protection provided by enclosures (IP code)
• IEC/EN 60068-2-6
Environmental influences: Test procedure - Test Fc: vibrate (sinusoidal).
• IEC/EN 60068-2-31
Environmental influences: Test procedure - Test Ec: shocks due to rough handling, primarily for
devices.
• EN 61010-1
Safety requirements for electrical equipment for measurement, control and laboratory use general requirements
• IEC/EN 61326
Emission in accordance with Class A requirements. Electromagnetic compatibility (EMC
requirements).
• NAMUR NE 21
Electromagnetic compatibility (EMC) of industrial process and laboratory control equipment
• NAMUR NE 32
Data retention in the event of a power failure in field and control instruments with
microprocessors
• NAMUR NE 43
Standardization of the signal level for the breakdown information of digital transmitters with
analog output signal.
• NAMUR NE 53
Software of field devices and signal-processing devices with digital electronics
• NAMUR NE 80
The application of the pressure equipment directive to process control devices
• NAMUR NE 105
Specifications for integrating fieldbus devices in engineering tools for field devices
• NAMUR NE 107
Self-monitoring and diagnosis of field devices
Endress+Hauser97
Page 98
Proline Promass F 100
• NAMUR NE 131
Requirements for field devices for standard applications
• NAMUR NE 132
Coriolis mass meter
• NACE MR0103
Materials resistant to sulfide stress cracking in corrosive petroleum refining environments.
• NACE MR0175/ISO 15156-1
Materials for use in H2S-containing Environments in Oil and Gas Production.
Ordering information
Detailed ordering information is available from the following sources:
• In the Product Configurator on the Endress+Hauser website: www.endress.com -> Click "Corporate"
-> Select your country -> Click "Products" -> Select the product using the filters and search field ->
Open product page -> The "Configure" button to the right of the product image opens the Product
Configurator.
• From your Endress+Hauser Sales Center: www.addresses.endress.com
Product Configurator - the tool for individual product configuration
• Up-to-the-minute configuration data
• Depending on the device: Direct input of measuring point-specific information such as
measuring range or operating language
• Automatic verification of exclusion criteria
• Automatic creation of the order code and its breakdown in PDF or Excel output format
• Ability to order directly in the Endress+Hauser Online Shop
Heartbeat Technology
Application packages
Many different application packages are available to enhance the functionality of the device. Such
packages might be needed to address safety aspects or specific application requirements.
The application packages can be ordered with the device or subsequently from Endress+Hauser.
Detailed information on the order code in question is available from your local Endress+Hauser sales
center or on the product page of the Endress+Hauser website: www.endress.com.
Detailed information on the application packages:
Special Documentation for the device → 102
PackageDescription
Heartbeat Verification
+Monitoring
Heartbeat Verification
Meets the requirement for traceable verification to DIN ISO 9001:2008 Chapter
7.6 a) "Control of monitoring and measuring equipment".
• Functional testing in the installed state without interrupting the process.
• Traceable verification results on request, including a report.
• Simple testing process via local operation or other operating interfaces.
• Clear measuring point assessment (pass/fail) with high test coverage within the
framework of manufacturer specifications.
• Extension of calibration intervals according to operator's risk assessment.
Heartbeat Monitoring
Continuously supplies data, which are characteristic of the measuring principle, to
an external condition monitoring system for the purpose of preventive
maintenance or process analysis. These data enable the operator to:
• Draw conclusions - using these data and other information - about the impact
process influences (such as corrosion, abrasion, buildup etc.) have on the
measuring performance over time.
• Schedule servicing in time.
• Monitor the process or product quality, e.g. gas pockets.
98Endress+Hauser
Page 99
Proline Promass F 100
Concentration
Special density
PackageDescription
ConcentrationCalculation and outputting of fluid concentrations
The measured density is converted to the concentration of a substance of a binary
mixture using the "Concentration" application package:
• Choice of predefined fluids (e.g. various sugar solutions, acids, alkalis, salts,
ethanol etc.)
• Common or user-defined units (°Brix, °Plato, % mass, % volume, mol/l etc.) for
standard applications.
• Concentration calculation from user-defined tables.
The measured values are output via the digital and analog outputs of the device.
PackageDescription
Special densityMany applications use density as a key measured value for monitoring quality or
controlling processes. The device measures the density of the fluid as standard and
makes this value available to the control system.
The "Special Density" application package offers high-precision density
measurement over a wide density and temperature range particularly for
applications subject to varying process conditions.
Accessories
Various accessories, which can be ordered with the device or subsequently from Endress+Hauser, are
available for the device. Detailed information on the order code in question is available from your
local Endress+Hauser sales center or on the product page of the Endress+Hauser website:
www.endress.com.
Device-specific accessoriesFor the sensor
AccessoriesDescription
Heating jacketIs used to stabilize the temperature of the fluids in the sensor.
Communication-specific
accessories
AccessoriesDescription
Commubox FXA195
HART
Commubox FXA291Connects Endress+Hauser field devices with a CDI interface (= Endress+Hauser
HART Loop Converter
HMX50
Wireless HART adapter
SWA70
Water, water vapor and other non-corrosive liquids are permitted for use as fluids.
If using oil as a heating medium, please consult with Endress+Hauser.
Heating jackets cannot be used with sensors fitted with a rupture disk.
For details, see Operating Instructions BA00132D
For intrinsically safe HART communication with FieldCare via the USB interface.
For details, see "Technical Information" TI00404F
Common Data Interface) and the USB port of a computer or laptop.
For details, see the "Technical Information" document TI405C/07
Is used to evaluate and convert dynamic HART process variables to analog current
signals or limit values.
For details, see "Technical Information" TI00429F and Operating Instructions
BA00371F
Is used for the wireless connection of field devices.
The WirelessHART adapter can be easily integrated into field devices and existing
infrastructures, offers data protection and transmission safety and can be operated
in parallel with other wireless networks with minimum cabling complexity.
For details, see Operating Instructions BA00061S
Endress+Hauser99
Page 100
Proline Promass F 100
Fieldgate FXA320Gateway for the remote monitoring of connected 4 to 20 mA measuring devices via
a Web browser.
For details, see "Technical Information" TI00025S and Operating Instructions
BA00053S
Fieldgate FXA520Gateway for the remote diagnostics and remote configuration of connected HART
measuring devices via a Web browser.
For details, see "Technical Information" TI00025S and Operating Instructions
BA00051S
Field Xpert SFX350Field Xpert SFX350 is a mobile computer for commissioning and maintenance. It
enables efficient device configuration and diagnostics for HART and FOUNDATION
Fieldbus devices and can be used in non-hazardous areas.
For details, see Operating Instructions BA01202S
Field Xpert SFX370Field Xpert SFX370 is a mobile computer for commissioning and maintenance. It
enables efficient device configuration and diagnostics for HART and FOUNDATION
Fieldbus devices and can be used in the non-hazardous area and in the hazardous
area.
For details, see Operating Instructions BA01202S
Service-specific accessories
AccessoriesDescription
ApplicatorSoftware for selecting and sizing Endress+Hauser measuring devices:
• Choice of measuring devices for industrial requirements
• Calculation of all the necessary data for identifying the optimum flowmeter: e.g.
nominal diameter, pressure loss, flow velocity and accuracy.
• Graphic illustration of the calculation results
• Determination of the partial order code, administration, documentation and
access to all project-related data and parameters over the entire life cycle of a
project.
Applicator is available:
• Via the Internet: https://wapps.endress.com/applicator
• As a downloadable DVD for local PC installation.
W@MW@M Life Cycle Management
Improved productivity with information at your fingertips. Data relevant to a plant
and its components is generated from the first stages of planning and during the
asset’s complete life cycle.
W@M Life Cycle Management is an open and flexible information platform with
online and on-site tools. Instant access for your staff to current, in-depth data
shortens your plant’s engineering time, speeds up procurement processes and
increases plant uptime.
Combined with the right services, W@M Life Cycle Management boosts
productivity in every phase. For more information, visit
www.endress.com/lifecyclemanagement
FieldCareFDT-based plant asset management tool from Endress+Hauser.
It can configure all smart field units in your system and helps you manage them. By
using the status information, it is also a simple but effective way of checking their
status and condition.
For details, see Operating Instructions BA00027S and BA00059S
DeviceCareTool for connecting and configuring Endress+Hauser field devices.
For details, see Innovation brochure IN01047S
Commubox FXA291Connects Endress+Hauser field devices with a CDI interface (= Endress+Hauser
Common Data Interface) and the USB port of a computer or laptop.
For details, see "Technical Information" TI00405C
100Endress+Hauser