Portable instrument for non-intrusive, quick ultrasonic flow measurement with clamp-on technology for all
types of piping
Features
• Precise bi-directional and highly dynamic flow measurement with the non-intrusive clamp-on technology
• High precision at fast and slow flow rates, high temperature and zero point stability
• Portable, easy-to-use flow transmitter with 2 flow channels, multiple inputs/outputs, an integrated data logger
with a serial interface
• Water and dust-tight (NEMA 4); resistant against oil,
many liquids and dirt
• Li-Ion battery provides up to 14 hours of measurement
operation
• Automatic loading of calibration data and transducer
detection for a fast and easy set-up (less than 5 min),
providing precise and long-term stable results
• User-friendly design
• Transducers available for a wide range of inner pipe
diameters () and fluid temperatures
()
• Probe for wall thickness measurement available
• Robust, water-tight (NEMA 4) transport case with comprehensive accessories
• HybridTrek automatically switches between transit time
and NoiseTrek mode of measurement when high particulate flows are encountered
• QuickFix for fast mounting of the flow transmitter in difficult conditions
FLUXUS supported by handle
Measurement with transducers mounted by mounting frames
and flow transmitter fixed to the pipe by the QuickFix pipe
mounting fixture
Applications
Designed for the following industries:
• Chemical industry
• Water and wastewater industry
• Oil and gas industry
• Cooling systems and air conditioners
• Facility management
• Aviation industry
Measurement equipment in transport case
TSFLUXUS_F601V1-5-1US_Lus, 2014-02-251
FLUXUS® F601Technical Specification
Table of Contents
Function ........................................................................................................................................................... 3
Calculation of Volumetric Flow Rate ................................................................................................................. 3
Number of Sound Paths.................................................................................................................................... 4
Technical Data .................................................................................................................................................. 6
Standard Scope of Supply ................................................................................................................................ 9
Connection of Adapters................................................................................................................................... 10
Example for the Equipment of a Transport Case ............................................................................................ 11
Transducer Order Code .................................................................................................................................. 13
Technical Data ................................................................................................................................................ 14
In order to measure the flow of a medium in a pipe, ultrasonic signals are used, employing the transit time difference principle. Ultrasonic signals are emitted by a transducer installed on the pipe and received by a second transducer. These signals are emitted alternately in the flow direction and against it.
As the medium in which the signals propagate is flowing, the transit time of the ultrasonic signals in the flow
direction is shorter than against the flow direction.
The transit time difference, ∆t, is measured and allows the flowmeter to determine the average flow velocity
along the propagation path of the ultrasonic signals. A flow profile correction is then performed in order to obtain the area averaged flow velocity, which is proportional to the volumetric flow rate.
Two integrated microprocessors control the entire measuring process. This allows the flowmeter to remove
disturbance signals, and to check each received ultrasonic wave for its validity which reduces noise.
HybridTrek
If the gaseous or solid content in the medium increases occasionally during measurement, a measurement
with the transit time difference principle is no longer possible. NoiseTrek mode will then be selected by the
flowmeter. This measurement method allows the flowmeter to achieve a stable measurement even with high
gaseous or solid content.
The transmitter can switch automatically between transit time and NoiseTrek mode without any changes to
the measurement setup.
Path of the ultrasonic signal Transit time difference ∆t
Calculation of Volumetric Flow Rate
= kRe . A . ka . ∆t/(2 . tfl)
where
=volumetric flow rate
k
=fluid mechanics calibration factor
Re
A=cross-sectional pipe area
k
=acoustical calibration factor
a
∆t=transit time difference
t
=transit time in the medium
fl
TSFLUXUS_F601V1-5-1US_Lus, 2014-02-253
FLUXUS® F601Technical Specification
a
a
a > 0
a < 0
Number of Sound Paths
The number of sound paths is the number of transits of the ultrasonic signal through the medium in the pipe.
Depending on the number of sound paths, the following methods of installation exist:
• reflect arrangement
The number of sound paths is even. Both of the transducers are mounted on the same side of the pipe.
Correct positioning of the transducers is easier.
• diagonal arrangement
The number of sound paths is odd. Both of the transducers are mounted on opposite sides of the pipe.
• direct mode
Diagonal mode with 1 sound path. This should be used in the case of a high signal attenuation by the medium, pipe or coatings.
The preferred method of installation depends on the application. While increasing the number of sound paths
increases the accuracy of the measurement, signal attenuation increases as well. The optimum number of
sound paths for the parameters of the application will be determined automatically by the transmitter.
As the transducers can be mounted with the transducer mounting fixture in reflect arrangement or diagonal arrangement, the number of sound paths can be adjusted optimally for the application.
.
a = transducer distance
Reflect arrangement, number of sound paths: 2
Diagonal arrangement, number of sound paths: 3
Direct mode, number of sound paths: 1Direct mode, number of sound paths: 1,
negative transducer distance
4TSFLUXUS_F601V1-5-1US_Lus, 2014-02-25
Technical SpecificationFLUXUS® F601
U L T R A S O N I C F L O W M E T E R
13
2
0
465
7
8
N E X T
Q
O N
Q -
Q +
M U X
D IS P
Q
O FF
L IG H T
E N T E R
C
O N
C H A N N E L A
C H A N N E L B
9
D IS P
E N T E R
3 x O F F
M O D E
B R K
B A TT E R Y
transducers
transmitter
RS232
outputs
inputs
power supply unit/battery charging unit
U L T R A S O N I C F L O W M E T E R
13
2
0
465
7
8
N E X T
Q
O N
Q -
Q +
M U X
D IS P
Q
O FF
L IG H T
E N T E R
C
O N
C H A N N E L A
C H A N N E L B
9
D IS P
E N T E R
3 x O F F
M O D E
B R K
B A TT E R Y
transducers
transmitter
temperature probe
temperature probe
supply line
return line
volumetric flow rate
supply temperature
return temperature
Typical Measurement Setup
Example of a measurement setup in reflect arrangement
TSFLUXUS_F601V1-5-1US_Lus, 2014-02-255
Example of a heat flow measurement
FLUXUS® F601Technical Specification
±1.6 % of reading ±0.03 ft/s
±1.2 % of reading ±0.03 ft/s
Flow Transmitter
Technical Data
FLUXUSF601
designportable
measurement
measurement principletransit time difference correlation principle,
flow velocity0.03 to 82 ft/s
repeatability0.15 % of reading ±0.03 ft/s
mediumall acoustically conductive liquids with < 10 % gaseous or solid content in volume (transit time difference
temperature compensationcorresponding to the recommendations in ANSI/ASME MFC-5.1-2011
accuracy
1
with standard calibration
with advanced calibration
(optional)
with field calibration
2
flow transmitter
power supply100 to 240 V/50 to 60 Hz (power supply unit),
batteryLi-Ion, 7.2 V/4.5 Ah
power consumption< 6 W
number of flow measuring
channels
signal attenuation0 to 100 s, adjustable
measuring cycle (1 channel) 100 to 1000 Hz
response time1 s (1 channel), option: 70 ms
housing materialPA, TPE, AutoTex, stainless steel
degree of protectionNEMA 4
dimensionssee dimensional drawing
weight4.2 lb
fixationQuickFix pipe mounting fixture
ambient temperature14 to 140 °F
display2 x 16 characters, dot matrix, backlight
menu languageEnglish, German, French, Dutch, Spanish
measuring functions
physical quantitiesvolumetric flow rate, mass flow rate, flow velocity,
totalizervolume, mass, optional: heat quantity
calculation functionsaverage, difference, sum
diagnostic functionssound speed, signal amplitude, SNR, SCNR, standard deviation of amplitudes and transit times
for transit time difference principle, reference conditions and v > 0.49 ft/s
2
reference uncertainty < 0.2 %
automatic NoiseTrek selection for measurements with high gaseous or solid content
principle)
±0.5 % of reading ±0.03 ft/s
10.5 to 15 V DC (socket at transmitter),
integrated battery
operating time (without outputs, inputs and backlight): > 14 h
2
heat flow (if temperature inputs are installed)
6TSFLUXUS_F601V1-5-1US_Lus, 2014-02-25
Technical SpecificationFLUXUS® F601
FLUXUSF601
communication
interfaceRS232/USB
serial data kit
software (all Windows™ versions)
- FluxData: download of measurement data, graphical presentation,
conversion to other formats (e.g. for Excel™)
- FluxKoef: creating medium data sets
- FluxSubstanceLoader: upload of medium data sets
сableRS232
adapterRS232 - USB
transport case
dimensions19.7 x 15.7 x 7.5 in
outputs
The outputs are galvanically isolated from the transmitter.
numbersee standard scope of supply on page 9, max. on request
accessoriesoutput adapter (if number of outputs > 4)
current output
range0/4 to 20 mA
accuracy0.1 % of reading ±15 μA
active outputR
passive outputU
< 200 Ω
ext
= 4 to 16 V, depending on R
ext
R
< 500 Ω
ext
ext
frequency output
range0 to 5 kHz
open collector24 V/4 mA
binary output
optorelay26 V/100 mA
binary output as alarm output
- functionslimit, change of flow direction or error
binary output as pulse output
- pulse value0.01 to 1000 units
- pulse width1 to 1 000 ms
inputs
The inputs are galvanically isolated from the transmitter.
numbersee standard scope of supply on page 9, max. 4
accessoriesinput adapter (if number of inputs > 2)
temperature input
typePt100/Pt1000
connection4-wire
range-238 to +1040 °F
resolution0.01 K
accuracy±0.01 % of reading ±0.03 K
current input
accuracy0.1 % of reading ±10 μA
passive inputR
- range-20 to +20 mA
= 50 Ω, Pi < 0.3 W
i
voltage input
range0 to 1 V
accuracy0.1 % of reading ±1 mV
internal resistanceR
= 1 MΩ
i
TSFLUXUS_F601V1-5-1US_Lus, 2014-02-257
FLUXUS® F601Technical Specification
F601
U L T R A S O N I C F L O W M E T E R
13
2
0
465
7
8
N E X T
Q
O N
Q -
Q +
M U X
D IS P
Q
O F F
L I G H T
E N T E R
C
O N
C H A N N E L A
C H A N N E L B
9
D IS P
E N T E R
3 x O F F
M O D E
B R K
B A T T E R Y
8.9
2.32
8.39
Dimensions
FLUXUS
in inch
8TSFLUXUS_F601V1-5-1US_Lus, 2014-02-25
Loading...
+ 17 hidden pages
You need points to download manuals.
1 point = 1 manual.
You can buy points or you can get point for every manual you upload.