Julabo F250, F500 Instruction Manual

1.951.4806-V4 11/17
www.julabo.com
Operating Manual
Recirculating Cooler
F250 F500 F1000
JULABO GmbH 77960 Seelbach / Germany Tel. +49 (0) 7823 / 51-0 Fax +49 (0) 7823 / 24 91
info.de@julabo.com
Temperature control devices for research and industry are developed,
9001
This product has been tested to the requirements of CAN/CSA-C22.2
requirements.
Congratulations!
You have made an excellent choice. JULABO thanks you for the trust you have placed in us. This operating manual has been designed to help you gain an understanding of the operation and possible applications of our circulators. For optimal utilization of all functions, we recommend that you thoroughly study this manual prior to beginning operation.
The JULABO Quality Management System
produced, and distributed according to the requirements of ISO and ISO 14001.Certificate Registration No. 01 100044846
For F250 device
No. 61010-1, second edition, including Amendment 1, or a later version of the same standard incorporating the same level of testing
Unpacking and inspecting
Unpack the circulator and accessories and inspect them for possible transport damage. Damage should be reported to the responsible carrier, railway, or postal authority, and a damage report should be requested. These instructions must be followed fully for us to guarantee our full support of your claim for protecting against loss from concealed damage. The form required for filing such a claim will be provided by the carrier.
Printed in Germany Changes without prior notification reserved
Important: keep original operating manual for future use
2 1.951.4806-V4
F250, F500, F1000
CONTENT
1. Intended use ................................................................................................................ 4
2. Operator responsibility – Safety recommendations ....................................................... 4
3. Handling......................................................................................................................... 5
3.1. Appropriate operation ............................................................................................ 5
3.2. Use ........................................................................................................................ 5
3.3. Disposal ................................................................................................................ 6
4. Technical specifications ................................................................................................. 7
4.1. F250 ...................................................................................................................... 7
4.2. F500 ...................................................................................................................... 8
4.3. Warning functions and safety installations .......................................................... 10
4.4. Materials of Construction of the wetted Parts ...................................................... 11
5. Safety Notes ........................................................................................................... 13
5.1. Description of the safety notes ............................................................................ 13
5.2. Explanation of other notes ................................................................................... 13
5.3. Safety instructions ............................................................................................... 14
6. Moving up and connect ................................................................................................ 15
6.1. Transportation and site conditions ....................................................................... 15
6.2. Tubing ................................................................................................................. 16
7. Operating controls and functional elements ................................................................. 18
7.1. F250 .................................................................................................................... 18
7.2. F500 .................................................................................................................... 20
8. Operating procedures .................................................................................................. 22
8.1. Bath fluids ........................................................................................................... 22
8.2. Power connection ................................................................................................ 23
8.3. Filling ................................................................................................................... 23
8.4. Switching on / Start - Stop ................................................................................... 24
8.5. Setting the temperatures ..................................................................................... 24
8.6. Timer function...................................................................................................... 25
8.6.1. Setting the time ................................................................................................ 25
8.6.2. Timer operation ................................................................................................ 25
8.7. AUTOSTART ON / OFF ...................................................................................... 26
9. Safety installations ....................................................................................................... 27
9.1. Excess temperature protection ............................................................................ 27
9.2. Low level protection ............................................................................................. 27
10. Troubleshooting guide / Error messages ..................................................................... 28
11. Cleaning / repairing the unit ......................................................................................... 30
11.1. Draining ............................................................................................................... 31
12. Warranty conditions ..................................................................................................... 32
12.1. EC-Declaration of Conformity .............................................................................. 33
1.951.4806-V4 3
Intended use
PID1
The recirculating coolers are operated via the splash-proof keypad.
technical products.
1. Intended use
JULABO recirculating coolers have been designed for temperature application to specific fluids. The pump connections can be used for cooling applications in an external circuit at a constant temperature.
The implemented microprocessor technology allows to set and to store the setpoint that can be indicated on the LED temperature display.
The PID temperature regulation is used to withdraw heat from the
bath fluid by means of the cooling machine and to automatically regulate the required need.
Health hazards caused by the bath fluid.
JULABO recirculating coolers are not conceived for direct temperature application to food and luxury articles or pharmaceutical and medico-
Direct temperature application means: Unprotected contact of the object with the bath medium (bath fluid).
2. Operator responsibility – Safety recommendations
The products of JULABO ensure safe operation when installed, operated, and maintained according to common safety regulations. This section explains the potential dangers that may arise when operating the circulator and also specifies the most important safety precautions to preclude these dangers as far as possible.
The operator is responsible for the qualification of the personnel operating the units. Make sure that the persons who operate the chillers, are trained in this work. The personnel operating the units should be regularly instructed about the dangers
involved with their job activities as well as measures to avert these dangers.
Make sure all persons tasked with operating, installing, and maintaining the unit have read
and understand the safety information and operating instructions.
When using hazardous materials or materials that could become hazardous, the circulator
may be operated only by persons who are absolutely familiar with these materials and the circulator. These persons must be fully aware of possible risks.
4 1.951.4806-V4
F250, F500, F1000
Contact: JULABO GmbH
77960 Seelbach / Germany
The unit is not for use in explosive atmosphere.
If you have any questions concerning the operation of your unit or the information in this manual, please contact us!
+49 7823 51-0 info.de@julabo.com
Eisenbahnstraße 45
+49 7823 2491 www.julabo.com
3. Handling
You have received a product designed for industrial use. Nevertheless, avoid strikes to the
housing, vibrations, damage to the operating-element panel (keypad, display), and contamination.
Make sure the product is checked for proper condition regularly (depending on the
conditions of use).
Regularly check (at least every 2 years) the proper condition of the mandatory, warning,
prohibition and safety labels.
Make sure that the mains power supply has low impedance to avoid any negative effects
on instruments being operated on the same mains.
This unit is designed for operation in a controlled electromagnetic environment. This
means that transmitting devices (e.g., cellular phones) should not be used in the immediate vicinity. Magnetic radiation may affect other devices with components sensitive to magnetic fields (e.g., monitors). We recommend maintaining a minimum distance of 1 m.
Permissible ambient temperature: max. 40 °C, min. 5 °C. Permissible relative humidity: 50% (40 °C). Do not store the unit in an aggressive atmosphere. Protect the unit from contamination. Do not expose the unit to sunlight.

3.1. Appropriate operation

Only qualified personnel is authorized to perform configuration, installation, maintenance and repairs of the circulator. Routine operation can also be carried out by untrained personnel who should however be instructed by trained personnel.
3.2. Use
Fire hazard!
The bath can be filled with flammable materials. Fire hazard! There might be chemical dangers depending on the bath medium used. Observe all warnings for the used materials (bath fluids) and the respective instructions (safety data sheets). Insufficient ventilation may result in the formation of explosive mixtures. Only use the unit in
1.951.4806-V2 5
Handling
or
Valid in EU countries
municipal waste is not permitted!
well ventilated areas. The unit is not for use in explosive atmosphere. Only use recommended materials (bath fluids). Only use non-acid and non corroding materials.
When using hazardous materials or materials that could become hazardous, the operator must affix the enclosed safety labels (1 + 2) to the front of the unit so they are highly visible:
1
2a
2b
Danger area. Attention! Observe instructions. (operating manual, safety data sheet)
Carefully read the user information prior to beginning operation. Scope: EU
Carefully read the user information prior to beginning operation. Scope: USA, NAFTA

3.3. Disposal

This unit contains the refrigerants R134a – at this time considered not to have any negative effects on the ozone layer. However, during the long operating period of the unit, disposal prescriptions may change. So only qualified personnel should take care of disposal.
See the current official journal of the European Union – WEEE directive. Directive of the European Parliament and of the Council on waste electrical and electronic equipment (WEEE). This directive requires electrical and electronic equipment marked with a crossed-out trash can to be disposed of separately in an environmentally friendly manner. Contact an authorized waste management company in your country. Disposal with household waste (unsorted waste) or similar collections of
6 1.951.4806-V4
4. Technical specifications
Recirculating Cooler

F250

Working temperature range
°C
-10 ... +40
Temperature stability
°C
±0.5
Temperature selection:
digital
via key pad
indication on LED-DISPLAY
Temperature indication:
LED-DISPLAY
Adjustment and display resolution
°C
0.1
Temperature control
PID 1
Temperature sensor
Pt 100
Excess temperature protection
85 °C - fixed value
Low liquid level protection
float switch
Circulating pump:
discharge, max.at 0 bar
l/min
15
pressure, max. at 0 Liters
bar
0.35
Filling level indicator
sight glass
Filling volume from ... to
Liters
1.7 ... 2.6
Dimensions (WxLxH)
cm
24x40x52
Weight
kg
27.0
Ambient temperature range
°C
5 ... 40
Return flow temperature max.
°C
80
Cooling compressor
1- stage / air cooled
Refrigerant
R134a
Cooling capacity
Medium: Mixture water-glycol
Mains power connection
V/ Hz
230 ±10 % / 50
Current draw (at 230 V)
A
3.0
Mains power connection
V/ Hz
230 ±10 % / 60
Current draw (at 230 V)
A
2.0
Mains power connection
V/ Hz
200 -5 %; +21 % / 50-60
Current draw (at 200 V / 50 Hz) (at 200 V / 60 Hz)
A A
2.0
2.0
Mains power connection
V/ Hz
115 ±10 % / 60
Current draw (at 115 V)
A
4.0
Mains power connection
V/ Hz
100 ±10 % / 50-60
Current draw (at 100 V / 50 Hz) (at 100 V / 60 Hz)
A A
5.0
5.0
4.1. F250
F250, F500, F1000
at 100 V / 60 Hz at 115 V / 60 Hz at 200 V / 60 Hz at 230 V / 60 Hz at 230 V / 50 Hz
at 100 V / 50 Hz at 200 V / 50 Hz
All measurements have been carried out at: rated voltage and frequency ambient temperature: 20 °C. Technical changes without prior notification reserved.
°C
W
°C
W
+20 +15 +10 +5 0 -5 -10 250 240 220 210 180 90 60
+20 +15 +10 +5 0 -5 -10 220 210 195 185 170 80 40
1.951.4806-V4 7
Technical specifications
Recirculating Cooler

F500

Working temperature range
°C
0 ... +40
Temperature stability
°C
±0.5
Temperature selection:
digital
via key pad
indication on LED-DISPLAY
Temperature indication:
LED-DISPLAY
Adjustment and display resolution
°C
0.1
Temperature control
PID 1
Temperature sensor
Pt 100
Excess temperature protection
85 °C - fixed value
Low liquid level protection
float switch
Circulating pump:
discharge, max.at 0 bar
l/min
24
pressure, max. at 0 Liters
bar
0.5
Filling level indicator
sight glass
Filling volume from ... to
Liters
5,0 ... 7,5
Dimensions (WxLxH)
cm
37,5x44x59
Weight
kg
34.0
Ambient temperature range
°C
5 ... 40
Return flow temperature max.
°C
80
Cooling compressor
1- stage / air cooled
Refrigerant
R134a
Cooling capacity
Mains power connection
V/ Hz
230 -10 %; +7 % / 50
Current draw (at 230 V)
A
3.0
Mains power connection
V/ Hz
230 ±10 % / 60
Current draw (at 230 V)
A
3.0
Mains power connection
V/ Hz
115±10 % / 60
Current draw (at 115 V)
A
6.0
Mains power connection
V/ Hz
100 -5 %; +10 % / 50-60
Current draw (at 100 V / 50 Hz) (at 100 V / 60 Hz)
A A
6.0
6.0
4.2. F500
at 230 V / 50 Hz Medium: Mixture water-glycol
All measurements have been carried out at: rated voltage and frequency ambient temperature: 20 °C Technical changes without prior notification reserved.
°C
W
+20 +10 +5 0 500 400 300 250
8 1.951.4806-V4
4.3. F1000
Recirculating Cooler

F1000

Working temperature range
°C
0 ... +40
Temperature stability
°C
±0.5
Temperature selection:
digital
via key pad
indication on LED-DISPLAY
Temperature indication:
LED-DISPLAY
Adjustment and display resolution
°C
0.1
Temperature control
PID 1
Temperature sensor
Pt 100
Excess temperature protection
85 °C - fixed value
Low liquid level protection
float switch
Circulating pump:
discharge, max.at 0 bar
l/min
23
pressure, max. at 0 Liters
bar
1.0
Filling level indicator
sight glass
Filling volume from ... to
Liter
7.0 ... 9.5
Dimensions (WxLxH)
cm
37.5x49x64
Weight
kg
46
Ambient temperature range
°C
5 ... 40
Return flow temperature max.
°C
80
Cooling compressor
1- stage / air cooled
Refrigerant
R134a
Cooling capacity
Mains power connection
V/ Hz
230 ±10 % / 50
Current draw (at 230 V / 50 Hz)
A
5.0
Mains power connection
V/ Hz
230 ±10 % / 60
Current draw (at 230 V / 60 Hz)
A
4.0
Mains power connection
V/ Hz
200 ±10 % / 50-60
Current draw (at 200 V / 50 Hz)
A
-
Current draw (at 200 V / 60 Hz)
A
-
Mains power connection
V/ Hz
115 ±10 % / 60
Current draw (at 115 V / 60 Hz)
A
9.0
F250, F500, F1000
at 230 V / 50 Hz Medium: Water-glycol
°C W
+20 +10 +5 0 1000 700 550 350
All measurements have been carried out at: rated voltage and frequency ambient temperature: 20 °C Technical changes without prior notification reserved.
1.951.4806-V4 9
Technical specifications
The unit is not for use in explosive atmosphere.

4.4. Warning functions and safety installations

Excess temperature protection 85 °C - fixed value Low liquid level protection float switch Alarm messages optical + audible (permanent) Overload protection for compressor and pump motor Classification according to DIN 12876-1 Class I
Environmental conditions according to IEC 61 010-1:
- Use only indoor.
- Altitude up to 2000 m - normal zero.
- Ambient temperature: +5 ... +40 °C
- Air humidity:
- Max. rel. humidity 80 % for temperatures up to +31 °C,
- linear decrease down to 50 % relative humidity at a temperature of +40 °C,
- max. permissible mains fluctuations, see Technical specifications.
The unit corresponds to Class I Overvoltage category II Pollution degree 2
Fire hazard!
Standards for interference resistance according to EN 61326-1 This unit is an ISM device classified in Group 1 (using high frequency for internal purposes) Class A (industrial and commercial range).
Information about the used refrigerants
The Regulation (EU) No. 517/2014 on fluorinated greenhouse gases applies to all systems which contain fluorinated refrigerants and replaces (EC) 842/2006. The aim of the Regulation is to protect the environment by reducing emissions of fluorinated greenhouse gases. Among other things it regulates the emission limits, use and recovery of these substances. It also contains requirements for operators of systems which require / contain these substances to function.
Under Regulation 517/2014, the operator of a system of this nature has the following duties:
The operator must ensure that the equipment is checked at regular intervals for leaks.
These intervals depend on the CO
refrigerant fill volume and type of refrigerant. The CO the model plate.
The operator undertakes to have maintenance, repair, service, recovery and recycling work
carried out by certified personnel who have been authorized by JULABO.
All such work must be documented. The operator must keep records and archive them for at
least five years. The records must be submitted to the relevant authority on request.
Refer to the text of the Regulation for further information.
equivalent of the system. This is calculated from the
2
equivalent of your system is shown on
2
10 1.951.4806-V4

4.5. Materials of Construction of the wetted Parts

F250
Designation
Material
Tube, inner diameter 8.0 x 2.0 mm
PVC
Sealings processed
PA
Bath, complete
1.4404, 1.4301, 1.4435
Profile sealing
Silicone, white
Filling pipe, above
PVC
Stopper
POM
O-ring
CR11-70 (Chloroprene rubber)
Motor mounting sheet
Motor plate
1.4301
Pump
1.4301, 1.4401, PPS (Rytone)
Sensors 2xPt 100
1.4571
Float switch
1.4301, PP
Barbed fittings
CuZn39Pb3 (nickel plated)
F500
Designation
Material
Tube
PVC
Sealings processed
PA
Bath, complete
1.4301, 1.4404
Sealing screw a.f. 13.0 x 11.0
1.4571
Profile sealing
Silicone, white
Filling pipe, above
PVC,grey
Stopper
POM
O-ring
CR11-70 (Chloroprene rubber)
Motor mounting sheet
Motor plate
1.4301/304H,1.4305/303
Pump
1.4301/304H, EPDM
Sensors 2xPt 100
1.4571
Float switch
1.4301/304, PP
Barbed fittings
1.4305/303
Sealing screw a.f. 13.0 x 11.0 1.4571
F250, F500, F1000
1.951.4806-V4 11
Loading...
+ 24 hidden pages