GE F-801 User Manual

GE OSMONICS
F-801 INTERNAL REVERSE OSMOSIS UNIT
Operation
and Maintenance Manual
Revision E
This Revision Supersedes P/n 12552 Rev. D
F-801 INTERNAL REVERSE OSMOSIS UNIT
Installation, Operation and Maintenance Manual
TABLE OF CONTENTS Page CHAPTER ONE: GENERAL INFORMATION 1-1
Section 1.1 The Manual 1-3
1.2 Safety Summary 1-3
1.3 Applications 1-4
1.4 Contraindications 1-4
1.5 Environmental Considerations 1-5
1.6 Theory of Operation 1-6
1.7 Fluid Components 1-7
1.8 Monitors and Controls 1-9
1.9 Optional Equipment 1-11
1.10
1.10.1
1.10.2
1.10.3
1.10.4
1.10.5
Specifications Inlet Water Requirements Flow Rates Electrical Requirements Membrane Performance Characteristics Environmental Requirements
1-12
1.11 Disposal 1-13
1.12 Electromagnetic Interference 1-13
1.13 Service Assistance 1-14
1.14 Return Goods Authorization (RGA) 1-15
1.15 Symbols and Abbreviations 1-16
CHAPTER TWO: INSTALLATION GUIDELINES AND INITIAL
2-1
SYSTEM START UP
Section 2.1 Installation Considerations 2-3
2.2 Installation Procedure 2-3
2.3 Product Water Analysis Procedure 2-5
CHAPTER THREE: SYSTEM OPERATION 3-1
Section 3.1 Daily Start-up Procedure 3-3
3.2 Bicarbonate Container Filling 3-6
3.3 Shut Down Procedure 3-6
A 27 February 2004
A
CHAPTER FOUR: ROUTINE MAINTENANCE 4-1
Section 4.1
4.1.1
Routine Maintenance of F-801 Internal RO System
4-3 4-3
Daily / Per Treatment Maintenance
4.1.2 Weekly Maintenance 4-5
4.1.3 Monthly Maintenance 4-5
4.1.4 Quarterly Maintenance 4-6
4.1.5 Annual Maintenance 4-6
4.2
4.2.1
Specific Maintenance Procedures and Instructions Motor Lubricant Procedure
4-7 4-7
4.2.2 Leak Repair 4-7
4.2.3 Pre-filter Replacement 4-9
4.3 F-801 Cleaning Indications 4-11
4.3.1 Pre-Cleaning Preparation 4-12
4.3.1.1 Pre-Cleaning Procedure 4-12
4.3.1.2 Cleaner Recirculation and Dwell Time 4-13
4.3.1.3 Post-Cleaning Rinse Procedure 4-14
4.3.2 BIOSAN Pre Cleaning Preparation 4-16
4.3.2.1 BIOSAN Cleaning Procedure 4-16
4.3.2.2 BIOSAN Recirculation and Dwell Time 4-17
4.3.2.3 BIOSAN Post Cleaning Rinse Procedure 4-18
4.4 F-801 Disinfection Indications 4-20
4.4.1 Formaldehyde Disinfection Preparation 4-20
4.4.1.1 Formaldehyde Pre-Disinfection Procedure 4-21
4.4.1.2 Formaldehyde Disinfection and Dwell Time 4-22
4.4.1.3 Formaldehyde Post-Disinfection Rinse Procedure 4-23
4.4.2 “RENALIN®” Disinfection Procedure
4-24
(Peracetic Acid and Peroxide-Based Compounds such as Renalin®, Minncare, & Peracidin™)
4.4.2.1 “RENALIN®” Pre-Disinfection Preparation 4-25
4.4.2.2 “RENALIN®” Disinfection and Dwell Time 4-26
4.4.2.3 “RENALIN®” Post-Disinfection Rinse Procedure 4-28
4.5 Product Water Culture 4-29
4.5.1 Product Water Culture Procedure 4-30
4.6 System Storage 4-31
4.6.1 Formaldehyde Storage Procedure 4-31
4.6.2 BIOSAN Storage Procedure 4-32
B 27 February 2004
B
CHAPTER FIVE: TROUBLESHOOTING 5-1
Table of Content 5-3 Section 5.1 RO Fails to Run 5-5
5.2 RO Runs Intermittently 5-5
5.3 Motor Runs Hot 5-6
5.4 Low Pump Pressure 5-6
5.5 Pressure Gauge Vibrates Erratically 5-7
5.6 Excess Product Pressure (too high) 5-7
5.7 Product Flow Too Low 5-7
5.8 Product Flow Too High 5-8
5.9 Membrane Product Pressure Too Low 5-9
5.10 Leaks 5-9
5.11 Low Percent Rejection 5-9
5.12 Rejection Display Fluctuates 5-10
5.13 TDS Value High 5-10
5.14 RO Fails to Draw Disinfectant 5-11
5.15 RO Pump Leaks 5-12
CHAPTER SIX: DRAWINGS 6-1
CHAPTER SEVEN: SPARE PARTS LIST 7-1
C 27 February 2004
C
This page intentionally left blank
D 27 February 2004
D
F-801 Internal RO System by GE Osmonics
F-801 Internal RO System
CHAPTER ONE:
GENERAL INFORMATION
12552e – 27 Feb 04 1 - 1 General Information
F-801 Internal RO System by GE Osmonics
This page intentionally left blank.
12552e – 27 Feb 04 1 - 2 General Information
F-801 Internal RO System by GE Osmonics
1.1 The Manual
This manual has been prepared to provide the operator with information and instructions regarding the installation, use, maintenance and troubleshooting of the GE Osmonics F-801 Internal Reverse Osmosis System.
CAUTION: When used as a medical device, Federal law restricts this device to sale by or on the order of a physician. Per CFR 801.109 (b)(1).
The manual has been written in narrative form supplemented with schematics and drawings for clarification. The operator can perform most procedures mentioned in this manual. Any exceptions will be clearly identified by a qualifying statement.
1.2 Safety Summary
Words in BOLD CAPITAL letters are used to identify labels on the device and key safety or qualifying statements. A list of all symbols and abbreviations is located at the end of this chapter. See Technote 120 for a glossary of relevant terms.
This safety summary does not contain all of the safety statements in the manual. Other safety statements are included within the manual text and are enhanced and defined as follows:
NOTE: Statements that provide further clarification.
CAUTION:
Statements identifying conditions or practices that could result in equipment or other property damage.
WARNING:
Statements identifying conditions or practices that could result in personal injury or loss of life.
READ THIS MANUAL:
Prior to operating or servicing this device, this manual must be read and understood. Keep this and other associated manuals for future reference and for new operators or qualified service personnel. A note sheet is provided at the end of each chapter for operators to make notations that may be valuable to other users.
USE PROPER POWER CONNECTIONS:
Use proper wiring and connection methods to satisfy hospital electrical codes.
DO NOT REMOVE COVERS OR PANELS:
To avoid electrical shock hazard, do not remove covers or panels when power is supplied to the device. Do not operate the device when covers or panels are removed.
12552e – 27 Feb 04 1 - 3 General Information
F-801 Internal RO System by GE Osmonics
USE PROPER WATER AND DRAIN CONNECTIONS:
Use proper water and drain connections to prevent contamination of the RO and danger to the patient. To ease confusion, the Quick Disconnect (QD) fittings have been color­coded. Both the male and female fittings are colored. The color coding is; Green = FEED water, Red = WASTE water, White = PERMEATE (Product) water.
SHOCK HAZARD:
Connect this device to a proper ground connection in accordance with the National Electrical Code. DO NOT under any circumstances remove the ground wire or ground prong from any power plug. DO NOT use an extension cord with this equipment. Keep the surface under the RO(s) free of liquids. Fix any leaks from the hemodialysis unit promptly.
DEVICE LABELING:
Do not under any circumstances; remove any Caution, Warning or any other descriptive labels from the devices until the conditions warranting the label are eliminated.
DO NOT OPERATE IN A FLAMMABLE ATMOSPHERE:
To avoid fire or explosion, do not operate this device in an explosive environment or near flammable anesthetics.
DO NOT TRY TO CATCH THE EQUIPMENT IF IT STARTS TO FALL OVER:
The overall assembly is weighty and can cause physical harm.
1.3 Applications
The F-801 Internal Reverse Osmosis (RO) System is designed to purify water by forcing water through a semi-permeable RO membrane. Water purified by reverse osmosis has had approximately 95% of the dissolved ions and approximately 99% of most other contaminants removed. The quality of the purified water, referred to as product water or permeate, depends on the quality of the feed water and normally is equal to or exceeds AAMI standards. Such water is suitable for hemodialysis applications.
1.4 Contraindications
WARNING:
GE Osmonics systems are not designed, sold, or intended for use in producing water for injection
Furthermore, GE Osmonics systems are not intended to be used outside of the device specifications and limitations, as outlined in this manual and other related materials.
12552e – 27 Feb 04 1 - 4 General Information
F-801 Internal RO System by GE Osmonics
1.5 Environmental Considerations
Prior to the installation of the F-801 RO System, it will be necessary to provide utilities and create an environment suitable for the trouble free operation of the RO system and its accessories.
POWER:
The F-801 operates on 115 VAC single-phase power. Histories of power failure, power surges, and low line voltages should be noted and reported to the manufacturer or their agent as they may create adverse conditions for the operation of equipment.
WATER:
Reverse osmosis systems require a reliable water source. The F-801 RO needs a minimum of 30-PSI pressure at a minimum flow of 1.6 gallons per minute. It is important that water is flowing at the design flow rate when testing water pressure rather than reading static pressure.
DRAIN:
A drain outlet is required by the RO system. The drain must have a minimum capacity of 1.6 gallons per RO per minute of continuous flow for as long as and as often as the RO is operating.
TEMPERATURE:
The F-801 RO is normally connected to a cold water supply. The performance specifications of the RO are based on 77°F (25°C) feed temperature. Each degree Fahrenheit the feed water temperature falls, the RO product flow decreases approximately 1.5%, and each degree Celsius drop, the product flow is reduced by 3% (See Technote 113 for the “Temperature Correction Factors” chart). In climates where the winter water temperature drops significantly below 77°F it may be necessary to temper (heat) the water.
WARNING:
The temperature of the tempered water should be monitored closely with a thermometer as high temperatures can damage the RO membranes and possibly harm the patient. Refer to the inlet water requirements (Section 1.10.1) for the maximum operating temperature for the F-801 RO.
CAUTION:
It is necessary to keep the RO from freezing to prevent serious damage to the RO unit.
NOISE:
The RO unit does not generate significant noise levels. Hard walls may reflect noise and will make the RO seem louder.
12552e – 27 Feb 04 1 - 5 General Information
F-801 Internal RO System by GE Osmonics
WATER PURITY:
Plumbing materials can significantly contribute to the contamination of the water. Metallic pipes (with the exception of 316 stainless steel) must be eliminated once the purification process has begun. Schedule 40 or 80 PVC pipes, polypropylene, PVDF, and other FDA recommended materials are suitable for most grades of purified water. Care must also be exercised in the choice of a thread sealant. Teflon tape is suitable for all threaded connections in this system.
WARNING:
Ordinary pipe dope must be avoided since it may leach objectionable and potentially dangerous impurities into the water.
1.6 Theory of Operation
The F-801 Portable Reverse Osmosis system (RO) purifies a given feed water (by way of reverse osmosis) for use in hemodialysis applications. The F-801 RO possesses monitoring for pressure, product water quality and membrane performance (percent rejection). The unit includes pre-filtration to remove particulates greater than 5 microns, which could clog orifices or damage components. The RO pump provides the pressure required to push water through the RO membrane. The fluid controls provide a means of managing flow rates and pressures. The following paragraphs detail the function of these fluid control components, which are used within the RO.
12552e – 27 Feb 04 1 - 6 General Information
F-801 Internal RO System by GE Osmonics
1.7 Fluid Components
Pre-filter:
From the feed water, the fluid path flows through the softener (optional) on through the carbon tank and then into the pre-filter, located in a blue housing mounted on the side of the hemodialysis unit. It is a cartridge type filter with a nominal filtration rating of five microns.
Pressure Adjust Regulator:
Used to adjust the product pressure. The pressure is displayed on the product pressure gauge, on the front of the RO.
High-Pressure Pump:
Provides up to a 200 PSI boost of the inlet water to the Reverse Osmosis Membrane Assembly (ROMA). The RO pump pressure is displayed on a 0 to 300 PSI gauge mounted outside the RO, below the product pressure gauge. The pressure is factory set at 150 PSI.
Reverse Osmosis Membrane Assembly (ROMA):
One is included in the F-801 RO. A ROMA is an assembly that includes the membrane housing, end caps, fittings, and a membrane. The thin film membrane will produce 600 (±15%) gallons per day of product water, depending on membrane condition and water temperature.
Waste Flow:
The Waste Flow is controlled by a length of 1/8" capillary tubing. The length of the tube is factory determined to correctly govern the wastewater flow to drain.
CAUTION:
This tube must not be shortened.
Dialyze/Flush Valve:
The Dialyze/Flush Valve is used to flush the membrane surface of rejected contaminants, and to accelerate the rinse out of cleaners or disinfectants from the RO. Some pump noise (chattering) is normal in the flush position. The RO is in "flush" from the full open (flush) to just before the fully closed (dialyze) position.
CAUTION:
Excessive “chattering” may be eliminated by slowly moving the valve from the full flush position towards “DIALYZE”.
Product Recirculation Check Valve:
Prevents over-pressuring of the product supply. When the product pressure exceeds the set point of the pressure regulator, the product recirculation check valve opens. Excess product water is then recirculated back to the inlet water flow.
12552e – 27 Feb 04 1 - 7 General Information
F-801 Internal RO System by GE Osmonics
Feed Port:
The Feed Port is a green colored quick-disconnection port for the introduction of feed water to the RO. The port has a check valve that closes when there is no hose connected to the fitting.
Product Port:
The Product Port is a white colored quick-disconnection port for the output of purified product water. The port has a check valve that closes when there is no hose connected to the fitting.
Waste Port:
The Waste Port is a red colored quick-disconnection port for the output of concentrate or waste to the drain. The port has a check valve that closes when there is no hose connected to the fitting.
12552e – 27 Feb 04 1 - 8 General Information
F-801 Internal RO System by GE Osmonics
1.8 Monitors and Controls
RO Power Switch:
The RO Power Switch is mounted on the back of the remote monitor module. The switch is used to turn the F-801 RO unit on and off.
Inlet Water Sensor:
Monitors the quality of the inlet water. Inlet quality is compared to the outlet (product) water quality and is displayed as "percent rejection" on the front panel.
Product Water Sensor:
The Product Water Sensor is used in conjunction with the inlet water sensor by the percent-rejection circuitry to calculate the membrane rejection rate. The rate of rejection is displayed from 70% - 99% in a 5 LED array on the front panel.
Total Dissolved Solids (TDS) Sensor:
Samples the product water quality via the sensor probe, and the final quality of the product water is indicated on a digital display on the front panel as TDS in mg/L and ppm. The monitor is temperature compensated to adjust the display value to changing water temperature.
Pressure Adjust Regulator:
The Pressure Adjust Regulator inside the RO is used to set the maximum product pressure needed to supply a dialysis machine. Adjusting this regulator will also result in small adjustments in the pump pressure.
Pre and Post filter Pressure Gauges:
The Pre-filter pressure gauges are mounted on top of the pre-filter housing and display the feed water pressure from zero to 100 PSI. The two gauges are used to determine the feed water pressure and when compared against each other to determine fouling of the filter (delta pressure).
Product Pressure Gauge:
The Product Pressure Gauge reads the product pressure from 0 to 100 PSI and is located on the RO module right side panel. Product pressure will read zero (or close to) when the product line is not connected to a dialysis unit and the product hose is connected to the product fitting and the valve is open. The Fresenius dialysis machine requires a minimum inlet pressure of 20 PSI.
Pump Pressure Gauge:
The Pump Pressure Gauge reads the pump pressure from 0 to 300 PSI and is located on the RO module right side panel. It is used for monitoring during operation and when adjusting the pump pressure, and is located on the front of the RO.
12552e – 27 Feb 04 1 - 9 General Information
F-801 Internal RO System by GE Osmonics
Dialyze/Flush Valve:
The Dialyze/Flush Valve is a 1/4-turn ball valve. When in the dialyze mode, the valve is in the closed position. When the valve is turned at any point towards flush, the valve is open. The closer to the flush point, the more the valve is open. When in flush, the valve bypasses the waste flow tube.
Percent Rejection Display:
The Percent Rejection Display is a 5 LED array that indicates how the RO membranes are performing. The lights correspond to 99%, 95%, 90%, 85%, and >70% rejection. The 99% light, when illuminated, indicates a need to evaluate the RO system performance. RO units do not normally perform in excess of 99%; verify this reading with an independent meter and take corrective action if indicated. Some RO membranes will achieve this level of performance under optimum conditions.
Product Water Total Dissolved Solids (TDS) Display:
The Product Water Total Dissolved Solids (TDS) digital display indicates product water purity in mg/L (ppm). It is temperature compensated to adjust the value of the display with changing water temperature for better accuracy.
Less Than 0.5 mg/L Check System Lamp:
The Less Than 0.5 mg/L Check System lamp illuminates when the TDS reading is
0.5 mg/L or less. This indicates a need to evaluate the RO system performance. RO units do not normally have this low product TDS. Verify this reading with an independent meter and take corrective action if indicated.
NOTE:
RO membranes may achieve this level of performance under optimum conditions (good feed water).
Alarm Set:
The Alarm Set switch allows the operator to set the TDS levels at which the alarm will sound. The alarm should be set at twice the initial TDS reading. If the set point is exceeded, an audible alarm will sound.
NOTE:
Choose the alarm set point that is closest to twice the initial installation TDS reading.
Alarm Mute:
The Alarm Mute switch will mute the audible alarm for 30 - 45 seconds. A red indicator lamp is displayed on the switch when the mute switch has been pressed.
12552e – 27 Feb 04 1 - 10 General Information
F-801 Internal RO System by GE Osmonics
1.9 Optional Equipment
GE Osmonics offers a number of optional items to adapt the equipment to meet specific needs. Options are available to improve the quality of the feedwater to the RO. Frequently requested options are briefly described in the following paragraphs. If more information is needed, or if other options are desired, please consult the factory. Refer to Chapter xx for catalog part numbers.
Carbon Filter Tanks remove chlorine/chloramine from the feed water, preventing damage to the RO membrane and injury to patients. The filter incorporates a bed of granular activated carbon. Two sizes of carbon tanks come in semi-automatic control valve and exchange head styles.
WARNING:
Carbon is not regenerated and must be replaced when exhausted.
WARNING:
A suitable carbon filter should always be present as pretreatment for the RO unit in order to prevent chlorine/chloramine damage to the thin film RO membrane and harm to the patient. GE Osmonics strongly recommends the use of two carbon tanks, used in a series configuration.
Softeners remove calcium, magnesium, and other scale producing contaminants from the feed water. Their removal prevents the build-up of scale on the membrane surface. A control valve drives the regeneration of the softener, and will remove the hardness ions with a brine solution and rinse the ion exchange bed afterwards. The control valve is normally governed by a semi-automatic time clock. There are two sizes of softener tanks available.
Multi-Media Filter Tanks are a multi layered filtration system that consists of various sized rocks as the media layers. The media are selected on the basis of their size and specific gravities. Capable of removing ninety-eight percent of all particles greater than ten microns, the multi-media filter is employed to remove suspended solids and particulates from the water. The use of the multi-media filter will extend the life of the pre-filter and RO membrane. Multi-media filters are not portable and not commonly used with portable RO systems.
Dealkalizers are used to lower the pH of alkaline feed water. Water with pH levels above 8.5 with chloramines present can reduce the performance of the carbon filters and thin-film membranes. A control valve drives the regeneration rinsing of the dealkalizer. The control valve is normally governed by a semi-automatic time clock. There are two sizes of dealkalizer tanks available.
12552e – 27 Feb 04 1 - 11 General Information
F-801 Internal RO System by GE Osmonics
1.10 Specifications:
1.10.1 Input Water Requirements
Minimum
Maximum
Input flow Rate 2.0 GPM N/A
Temperature (water) 10°C (50°F) 32°C (90°F)
NOTE: Operating at lower temperatures may make it impossible for the F-801 RO to produce adequate permeate for the dialysis machine.
pH (operating) 5.5 10.5 TF Membranes
pH w/chloramines present (operating) 5.5 8.5 TF Membranes
pH (cleaning) 2.0 11.0 TF Membranes
Pressure, dynamic 30 PSI 100 PSI
1.10.2 Flow Rates
Product Flow based on the following conditions: New Membranes Operating Pressure: 150 PSI Product Pressure: 1 PSI Feed Water Temperature: 77°F (25°C)
LPM GPD LPD
GPM .41 1.55 588 2225.8
Membrane flux may vary ±15% Waste Flow: 2000ml/min (±10%)
1.10.3 Electrical Requirements
Motor 1/3 HP
Voltage 115 V~
Hertz 60
Amps 10
Phase single
Power Cord (length) 8 ft (2.4 m)
12552e – 27 Feb 04 1 - 12 General Information
F-801 Internal RO System by GE Osmonics
1.10.4 Membrane Performance Characteristics
Salt rejection (TF Membranes): 95% minimum at initial testing at factory.
Test Solution: Inlet water 1000 PPM NaCl Conductivity: 2000 micro mhos-cm Operating Pressure: 200 PSI Product Back pressure: 0 PSI
1.10.5 Environmental Requirement
Minimum Maximum
Ambient Temperature 4°C (39°F) 32°C (90°F)
Storage Temperature 2°C (36°F) 32°C (90°F)
Altitude N/A 10,000 feet
CAUTION:
This equipment must not be allowed to freeze. Irreparable harm to various components, including RO membranes may result.
NOTE: The user/operator should recognize that moisture can be caused by condensation and is not necessarily an equipment leak. This equipment will function in the presence of condensation.
1.11 Disposal
Disposal of this product or parts must be carried out according with local disposal codes.
1.12 Electromagnetic Interference
This equipment can generate, uses and can radiate radio frequency energy and, if not installed and used in accordance with these instructions, may cause harmful interference to other devices in the vicinity. However, there is no guarantee that interference will not occur in a particular installation. If this equipment does cause harmful interference with other devices, which can be determined by turning the equipment off and on, the user is encouraged to try to correct the interference by one or more of the following measures:
Reorient or relocate the receiving device
Increase the separation between the equipment
Connect the equipment into an outlet on a circuit different from that to
which the other device(s) is connected
Consult the field service technician or manufacturer for help.
12552e – 27 Feb 04 1 - 13 General Information
F-801 Internal RO System by GE Osmonics
1.13 Service Assistance
If service assistance is required, please take the following steps:
Consult the “Troubleshooting” section of this manual (Chapter Five). If the problem cannot be identified and corrected by any of the procedures found in that section, then...
Contact your Facility Equipment Technician. If the technician is unable to help then…
Call GE Osmonics Technical Service Department at (800) 633-3080 or (206) 395-2200. Technicians are available for all calls between 7:00 a.m. and 4:30 p.m. PST, Monday through Friday. Technicians are also available at other times for emergency calls only. Product consultants will be on hand to discuss the problem with you and endeavor to rectify it over the phone. If the problem appears to be of a more serious nature, you will be given instructions regarding the action to be taken. Prior to making the phone call, you must be prepared to answer two questions:
1. What RO do you have, i.e. F-801
2. What is the serial number of your RO? (Label, found on back of Remote Monitor.)
RO SERIAL NUMBER: ______________________
12552e – 27 Feb 04 1 - 14 General Information
F-801 Internal RO System by GE Osmonics
1.14 Return Goods Authorization (RGA) Procedure
If you wish to return goods for, warranty evaluation and/or credit please have your original sales order, invoice and device serial number available when you call GE Osmonics. Call GE Osmonics at (800)-633-3080 and request Technical Service. A GE Osmonics Technical Service representative will provide instructions and a return authorization number, which needs to be clearly written on the outside of the box used to ship your materials. All equipment must be shipped to GE Osmonics with the freight prepaid by the customer. Call our Customer Service Center with any questions or issues concerning freight claims and a representative will discuss your situation.
All materials to be returned must be rendered into a non-hazardous condition prior to shipping.
12552e – 27 Feb 04 1 - 15 General Information
F-801 Internal RO System by GE Osmonics
1.15 Symbols and Abbreviations
“RENALIN®” Refers to Peracetic Acid and Peroxide-based Compounds, such as
Renalin®, Minncare™, Peracidin™ (20% peroxide, 4% Peracetic Acid)
ASTM American Society for Testing and Materials
AAMI Association for the Advancement of Medical Instrumentation
CA Cellulose Acetate
C Celsius
CSA Canadian Standards Association
cc Cubic Centimeters
cf Cubic Foot (feet)
cfu/ml. Colony Forming Units per milliliter
cm Centimeters
DI Deionization
F Fahrenheit
FDA Food and Drug Administration
Ft Foot (feet)
GPD Gallons Per Day
GPM Gallons Per Minute
GHT Garden Hose Thread
gr Grains
GAC Granular Activated Carbon
HZ Hertz
HP Horse Power
lbs. Pounds
LED Light Emitting Diode
LPD Liters Per Day
LPM Liters Per Minute
Kohm One Thousand Ohms
m. Meter
mg/L. Milligrams Per Liter
12552e – 27 Feb 04 1 - 16 General Information
F-801 Internal RO System by GE Osmonics
ml. Milliliters
MNPT Male National Pipe Thread
MSDS Material Safety Data Sheet
Megohm/cm Million Ohms/Centimeter
mw Molecular Weight
NaCl Sodium Chloride
N/A Not Applicable
NPT National Pipe Thread
NSF National Sanitation Foundation
PA Polyamide
P/N Part Number
ppm Parts Per Million
psi Pounds Per Square Inch
PVC Polyvinyl Chloride
QD Quick Disconnect
RO Reverse Osmosis
ROMA Reverse Osmosis Membrane Assembly
TB Terminal Barrier Strip: “TB2” (Example: “TB2-1” refers to the
terminal marked number 1 on the Power Supply Terminal Barrier Strip)
TF Thin Film
TDS Total Dissolved Solids
UL Underwriters Laboratory
VAC Volts Alternating Current
WFI Water for Injection
WT Weight
~ Volts Alternating Current
12552e – 27 Feb 04 1 - 17 General Information
NOTES:
F-801 Internal RO System by GE Osmonics
12552e – 27 Feb 04 1 - 18 General Information
F-801 Internal RO System by GE Osmonics
F-801 Internal RO System
CHAPTER 2:
INSTALLATION GUIDELINES AND INITIAL SYSTEM START-UP
12552e - 27 Feb 04 2 - 1 Installation Guidelines and Instructions
F-801 Internal RO System by GE Osmonics
This page intentionally left blank.
12552e - 27 Feb 04 2 - 2 Installation Guidelines and Instructions
F-801 Internal RO System by GE Osmonics
2.1 Installation Considerations
1. The RO unit requires 115 VAC power for operation and consumes 10 amps of current.
2. The RO unit requires nominally 1.6 GPM water at 30 PSI. Warmer feed water (up to 25°) will produce more permeate, and will consume more water. The water may also require some pretreatment to ensure maximum performance and suitability. GE Osmonics recommends that feed water be chlorine/chloramine free and solids above 5 microns in size be filtered out.
3. A drain with a minimum capacity of 1.6 GPM is required for the RO unit.
4. Back flow prevention may be required in accordance with local codes.
5. Special connections may be needed to connect to water supply. Connections vary with different facilities.
2.2 Installation Procedures:
CAUTION:
If a carbon tank is present on the RO system, the carbon tank must be completely rinsed of all carbon fines and ash before connecting to the RO unit. Otherwise, irreparable damage to the TF membrane will result (refer to Carbon Manual).
WARNING:
Do not rinse carbon tanks until they are ready to be used in service. Carbon tanks that have been wetted and allowed to sit unused will grow bacterial.
WARNING:
The ROMA is stored with 1 wt.% sodium metabisulfite from the factory. Follow all Material Safety Data Sheet (MSDS) guidelines that are included with the RO.
CAUTION:
Supply water with a pH above 8.5 and containing chloramines will adversely affect the operation of the RO membrane.
1. For installation of the pretreatment cart and RO to the dialysis machine cabinet, and all tubing connections (except product to use), refer to the Installation Drawing section of the system manual, or contact GE Osmonics for Drawing p/n 1239784.
2. Connect the product to use line to the product line rinse extension (GE Osmonics p/n 12393) and route to drain.
3. Connect the feed water line to water supply and route the RO waste line to drain.
4. Turn the water supply on and allow the water to flow through the system to drain for 3-5 minutes.
12552e - 27 Feb 04 2 - 3 Installation Guidelines and Instructions
F-801 Internal RO System by GE Osmonics
NOTE:
There will be little or no flow through the product line at this time.
5. Make sure that the RO power switch is in the OFF position. Plug the power cord into a 115 VAC grounded outlet. Do not use an extension cord or an adapter.
6. Ensure the dialyze/flush valve is in the dialyze position.
7. Turn the RO power switch on and run the system for 30 minutes. Observe displays and note stabilization of rejection and TDS.
8. After a minimum of 3 hours run time, disinfect and rinse the RO unit with an approved disinfectant before using for hemodialysis treatment (refer to Chapter Four for instructions on disinfection).
WARNING:
After disinfection, the product water must be verified free of residual disinfectant.
WARNING:
Ensure that TDS and membrane performance (% rejection) are normal for the regions water quality. Consult your facility physician. An AAMI analysis and bacteria test of the product water must be obtained and interpreted before using the RO unit for hemodialysis treatment (refer to the Product Water Analysis procedure and the Product Water Culture Procedure).
CAUTION:
New RO units with thin film (TF) membranes must be rinsed to drain with normal pressures for a minimum of three (3) hours before exposure to disinfectants such as Renalin and a minimum of six (6) hours before exposure to formaldehyde.
NOTE:
Ideally, rejection will be 95% or greater. If less than 95%, extend rinse period an additional 30 minutes.
9. Close the product valve and remove the product rinse extension line. Connect product line to inlet line on the dialysis machine. Open the product valve.
10. Start the dialysis machine into a rinse cycle.
11. Set the product water TDS alarm limit switch as prescribed by the physician. The switch is normally set to twice the initial TDS reading (or “next higher” setting) at start-up.
WARNING:
The TDS alarm limit should not be moved from this setting unless instructed by a physician.
12552e - 27 Feb 04 2 - 4 Installation Guidelines and Instructions
F-801 Internal RO System by GE Osmonics
12. Perform the following checks:
Pump pressure gauge: between 140 – 160 PSI Pre-filter pressure differential: Both above 30 PSI, no greater difference
than 8 PSI above clean filter readings.
Product pressure: between 35-40 PSI. Percent rejection: greater than 95%, continue to rinse if lower. TDS at normal for region.
13. The RO product may be used in conjunction with hemodialysis treatment equipment only after obtaining test results which validate compliance with excepted standards.
AAMI Product Water Panel Microbiological Culture (colony count) results Endotoxin (pyrogen) level testing (if applicable)
14. The F-801 RO is now ready for use.
2.3 Product Water Analysis Procedure:
In order to ensure that the RO is performing at or above AAMI (Association for the Advancement of Medical Instrumentation) Standards, it is necessary to perform an AAMI Water Quality Analysis test. For new units, the RO must be operated (rinsed to drain) for at least two hours before taking a sample, to ensure that the membranes are free of preservative or other contaminants which would invalidate the test.
WARNING:
There are many documented clinical risks in using less than AAMI Standard product water (refer to Technote 103, "AAMI Hemodialysis Water Quality Standards").
NOTE:
Follow the procedures recommended by your water quality-testing laboratory if they vary from the following steps.
Materials Required: Produce water sampling kit, sterile gauze, tape, and container (such as a small bucket).
1. Ensure that all pre-RO devices are operating correctly and are performance verified by the appropriate testing.
2. Connect the RO unit to the required utilities (feed water, drain, power).
3. With the dialysis machine on, operate the RO unit with product and waste flows to drain for at least 15 minutes after the product water quality has stabilized before drawing the water sample.
12552e - 27 Feb 04 2 - 5 Installation Guidelines and Instructions
F-801 Internal RO System by GE Osmonics
4. Close the product valve on the F-801, and disconnect the product line from the inlet water line of the dialysis machine. Tape a sterile gauze around the opening of the inlet water line to the dialysis machine.
5. Hold the product line over the container and run product water for 60 seconds.
6. Aseptically remove the lid from the sample bottle and rinse the bottle twice with RO product water.
7. Fill the bottle completely with RO product water and cap it immediately.
8. Fill out the laboratory requisition form. Ensure the requisition form has, at a minimum, information to identify the facility or location of the RO, the serial number or facility equipment number to ensure which RO was tested and name, phone number and address of whom to contact.
9. Place the bottle and the completed requisition in the pre-addressed mailer and mail it directly to the laboratory.
NOTE:
Total chlorine must be measured on-site. Tests for Chlorine/Chloramine (total chlorine) are not typically reported on the AAMI Analysis. Chlorine and chloramine dissipate from the water sample over time and reported results will be inaccurate.
12552e - 27 Feb 04 2 - 6 Installation Guidelines and Instructions