2103870.03 GB/D 01/2006
2103870.03 GB/D 01/2006
11WIKA Operating instructions / Betriebsanleitung F-20, F-21
10 WIKA Operating instructions / Betriebsanleitung F-20, F-21
7. Starting, operation
GB
7. Starting, operation
GB
Functional test
n Open pressure connections only after the system is without pressure!
n Observe the ambient and working conditions outlined in section 7
Technical data.
n Please make sure that the pressure transmitter is only used within the overload
threshold limit at all times!
Warn ing
When touching the pressure transmitter, keep in mind that the surfaces of the
instrument components might get hot during operation.
Caution
The output signal must be proportional to the pressure. If not, this might point to
a damage of the diaphragm. In that case refer to chapter 9 Trouble shooting.
When designing your plant, take into account that the stated values (e.g. burst pressure, over
pressure safety) apply depending on the material, thread and sealing element used.
Specifications
Model F-20, F-21
Power supply U
B
UB in DC V
10 < U
B
≤
30 (11 ... 30 with sig nal output 4 ... 20 mA,
14 ... 30 with signal outp ut 0 ... 10 V)
Signal output and RA in Ohm
4 ... 20 mA, 2-wire R
A
≤ (UB - 11 V) / 0,02 A
maximum load R
A
0 ... 20 mA, 3-wire RA ≤ (UB - 3 V) / 0,02 A
{0 ... 5 V, 3-wire} RA > 5 000 {0 ... 10 V, 3-wire} RA > 10 000
Test circuit signal / max. load R
A
Only for instrume nts with 4 ... 20 mA signal output;
RA < 15 Ohm with 20mA
Adjustability zero/span % ± 10 using potentiom eters inside the instrument
Response time (10 ... 90 %) ms
≤
1
Dielectric strength DC V 500
Accuracy % of span
≤
0.25 {0.125}
4)
(BFSL)
% of span
≤
0.5 {0.25}
4) 5)
4)
Accuracy { } for pressure ranges ≥ 0.25 bar
5)
Including non-linearity, hysteresis, non-repeatability, zero point and full scale error
(corresponds to error of measurement per IEC 61298-2).
Adjusted in vertical mounting position with lower pressure connection.
Non-linearity % of span
≤
0.2 (BFSL) according to IEC 1298-2
1-year stability % of span
≤
0.2 (at reference conditions)
Permissible temperature of
n
Medium
6) *)
-30 ... +100 °C {-40 ... +125 °C 7)} -22 ... +212°F {-40 ...+ 257°F 7)}
n
Ambience
6)
-20 ... + 80 °C {-30 ... +105 °C} -4 ... +176 °F {-22 ... +22 1°F}
n
Storage
6)
-40 ... +100 °C -40 ... +212 °F
6)
Also complies with EN 50178, Tab. 7, Operation (C) 4K4H, Storage (D) 1K4, Transport (E) 2K3
7)
Response time F-20: ≤ 10 ms
at medium temperatures below -30 °C for pressure
ranges up to 25 bar
Response time F-21: ≤ 10 ms at medium temperatures below -30 °C (-22 °F)
Specifications
Model F-20, F-21
Compensated temperature range 0 ... + 80 °C +32 ... +176 °F
Temperature coefficients within
compensated temp range
n
Mean TC of zero
% of span / 10 K
≤
0.2 (< 0.4 for pressure range 0 ... 0.1 and 0 ... 0.16 bar)
n
Mean TC of range % of span / 10 K
≤
0.2
a
- conformity
89/336/EEC interferenc e emission and immunity see
EN 61 326, interfere nce emission limit class A and B
97/23/EG Pressure eq uipment directive (Module H)
Shock resistance g 600 according to IEC 60068-2-27 (mechanical shock)
Vibration resistance g 10 acco rding to IEC 60068-2-6 (vibr ation under resonance)
Wiring protection Protected against reverse polarity and and sho rt circuiting on the
instrument side
Mass kg Approx. 0.35
*) In an oxygen version model F-21 is not available. In an oxygen version model F-20 is only available in gauge pressure ranges
≥
0.25 bar with media temperatures between -20 … +60 °C / -4 … +140 °F and using stainless steel or Elgiloy
®
wetted parts.
{ } Items in curved brackets ar e optional extras for additional price.