24.Setting Items list ..............................................................161
25.Statement of Conformation to EC Directives ................163
Per Reading This Instruction Manual
Per Reading This Instruction Manual
Since various names on the F395 are determined by assuming the X-axis sensor
displacement input, matters relating to the X-axis input may be expressed as
displacement - depending on the description in this instruction manual.
1
1. Outline of the F395
Sampling of waveform variation with
displacement for going and return,
detection of points in a waveform such
as an inflection point, and judgment of
differences between going and return
at a given point
(Hysteresis / Hysteresis 2 function);
Connection (P.21 )
Calibration (P.31 )
Settings Relating to Indicated
Value (P.38 )
Reading of Waveform (P.92 )
Measurement Operation
Functional Settings (P.40 )
Multi-Hold, (P.47 )
etc.
Waveform Comparison /
waveform and Displacement
Comparison, (P.68 )
etc.
The procedures from waveform display to uses of the Multi-Hold function and Waveform Comparison
function are as follows:
The Multi-Hold / Hysteresis / Hysteresis 2 / Waveform Comparison / Waveform and
Displacement Comparison functions cannot be used simultaneously.
Carefully read the explanations of "7. Measurement Operation Functional Settings" P40,
"10. Multi-Hold Mode" P47, "11. Hysteresis Mode" P64, "12. Hysteresis 2" P65, "13.
Waveform Comparison / Waveform and Displacement Comparison Mode" P68 and select
function(s) appropriate for the equipment application you use.
Detection of points in a waveform such
as an inflection point
(Multi-Hold function);
Basically, the waveform between the START signal and STOP
signal is graphically drawn.
(Conditional. See "9. Method of Starting Measurement" P43.)
START
STOP
Hysteresis (P.64 )
Hysteresis 2 (P.65 )
etc.
Check to see if the waveform variation
with time and displacement is within a
certain width
(Waveform Comparison / Waveform
and Displacement Comparison
function).
Waveform display of signals from a
strain gauge type sensor;
Strain gauge type sensor
Request
1-1. What the F395 Can Do:
1.Outline of the F395
2
1.Outline of the F395
1-2. Waveform Drawing Procedures
The F395 can provide waveform displays of signals from a strain gauge type sensor and
displacement sensor. The waveform display procedures are as follows.
(See "7-4. X/Y-axis Setting by Operation Mode and Input CH" P41.):
1. Connect a strain gauge type sensor (and displacement sensor) with the F395.
(See "4-1-1. Analog Input / Output Connection" P21.)
2. Perform calibration. (See "5. Methods of Calibration" P31.)
3. Set each parameter for waveform display.
(See "9. Method of Starting Measurement" P43, and "16-2-9. Level Axis Select /
4. Check to see if a waveform is displayed for the set conditions.
1-3. Multi-Hold Procedures
In the Multi-Hold mode, necessary points are detected in the displayed waveform to
make a judgment such as a High/Low Limit Comparison. Up to 32 channels of Hold
types, High and Low Limits, etc., can be stored, which can be switched by external
signals. The procedures for using the Multi-Hold function are as follows:
1. Check to see if a waveform is displayed in accordance with the waveform drawing
procedures. (See "1-2. Waveform Drawing Procedures" P3.)
2. Select a hold function so that necessary points in the waveform can be held. (See "10-
3. Hold Functions" P48.)
3. Carry out Multi-Hold settings. (See "16-2. Each Setting Item" P103.)
4. Select the channel number(s) you want to control externally. (See "4-3. Control
Connector Connection" P25.)
5. Input T/H and H/M signals in accordance with the selected hold function, and check to
see if operation is performed as set. (See "10-3. Hold Functions" P48.)
3
1-4. Hysteresis Procedures
In the Hysteresis mode, go and return waveform variations with displacement are
sampled, and necessary points in the displayed waveforms are detected to perform
controls such as a High/Low Limit Comparison.
This function is almost the same as the Multi-Hold function except for the return
waveform sampling function by hysteresis interval (See "16-2-14. Hysteresis Interval"
P119.), displacement on the X-axis, and the maximum sampling rate of 2kHz.
1. Check to see if a waveform is displayed in accordance with the waveform drawing
procedures. (See "1-2. Waveform Drawing Procedures" P3.)
2. Select a hold function such that necessary points in the waveform can be held. (See
"10-3. Hold Functions" P48.)
3. Carry out Hysteresis settings. (See "11-2. Setting and Operating Method" P64.)
4. Select the channel number(s) you want to control externally. (See "4-3. Control
Connector Connection" P25.)
1.Outline of the F395
5. Input T/H and H/M signals in accordance with the selected hold function, and check to
see if operation is performed as set. (See "10-3. Hold Functions" P48.)
1-5. Hysteresis 2 Procedures
The Hysteresis 2 mode is the same as the Hysteresis mode except for the hold operation.
At the point of displacement where a pulse is input, High/Low Limit Comparison of the
go measured value, and High/Low Limit Comparison of the difference between the go
and return measured values are performed.
1. Check to see if a waveform is displayed in accordance with the waveform drawing
procedures. (See "1-2. Waveform Drawing Procedures" P3.)
2. Carry out Hysteresis 2 settings. (See "12-2. Setting and Operating Method" P65.)
3. Select the channel number(s) you want to control externally.
(See "4-3. Control Connector Connection" P25.)
4. Provide a load to the strain gauge type sensor / displacement sensor, input START and
HOLD 1 ~ 3 signals (See "4-3. Control Connector Connection" P25.), and check to
see if operation is performed as set.
4
1.Outline of the F395
1-6. Waveform Comparison /
Waveform and Displacement Comparison Procedures
In the Waveform Comparison / Waveform and Displacement Comparison mode,
waveform variations with time / displacement are sampled, and a High/Low Limit
Comparison is performed based on the sampled waveform.
A sequential comparison can be performed with respect to dynamic waveform
variations.
The procedures for performing a Waveform Comparison / Waveform and Displacement
Comparison are as follows:
1. Check to see if a waveform is displayed in accordance with the waveform drawing
procedures. (See "1-2. Waveform Drawing Procedures" P3.)
2. Carry out Waveform Comparison / Waveform and Displacement Comparison settings.
(See "13-2. Setting and Operating Method" P68.)
3. Select the channel number(s) you want to control externally.
(See "4-3. Control Connector Connection" P25.)
4. Provide a load to the strain gauge type sensor (displacement sensor), input START and
other signals, and check to see if operation is performed as set.
1-7. Operation Mode at a Glance
Operation
mode
Multi-Hold
Hysteresis
Waveform
Comparison
Waveform and
Displacement
Comparison
Usable X-axisMaximum
Time
Sensor
○××
×○○
○××
×○○
Pulse
sampling
rate
4kHz9 points for all types
2kHz9 points for all types
4kHz1 point for three types
2kHz1 point for three types
Hold
Waveform
comparison
××
×○
○×
○×
Capability
of return
5
Hysteresis 2
×○○
2kHz3 points for one type
×○
1-8. Standard Interface
Status, set values, indicated values, waveform data, etc., can be sent to your PC through
the RS-232C serial interface. Set values and commands can also be sent from the PC to
F395. (See "18-2. RS-232C Interface" P133.)
Also, external equipment such as a printer and large display manufactured by us can be
used by using the SI/F 2-wire serial interface. (See "18-1. SI/F 2-wire Serial Interface"
P132.)
1-9. Waveform Reading
Measured values at sampling points for the latest sampled waveform and NG-
measurement waveforms (rejected waveform reading: four waveforms at maximum) can
1.Outline of the F395
be read.
(See "15-1. Waveform Reading" P92, "15-2. Rejected Waveform Reading" P96, and "15-
3. Rejected Waveform Clear" P97.)
6
2.Name and Function of Each Part
Touch panel type
color liquid
crystal display
Power lamp
Shifts to the
2. High and Low Limit
5. Lithium battery
6. Status display
3. X-axis (displacement)
4. Displacement High and
[Main screen]
1. Y-axis display
Performs Digital Zero.
ⅰ
ⅱ
ⅲ
ⅳ
7. Set Operation
8. Control Signal
9. Display of
In the Pulse Hold / Hysteresis 2 mode, the main screen layout is changed to
each specific one.
(See "10-3-10. Pulse Hold" P62, and "12. Hysteresis 2" P65.)
Shifts to the Hold Point
Set Values
section
display section
Low Limit Set Values
alarm
section
Mode display
display
High and Low
Limit Set Values
Display screen.
Mode Setting
Selection screen.
BALM
2. Name and Function of Each Part
2-1. Front Pane
2-1-1. Touch Panel Type Color Liquid Crystal Display
Indicated values and graphs are displayed, and various items are set on this touch panel
type color liquid crystal display.
7
[Setting screen]
setting screen with the key displayed in his place.
Returns to the
Setting item
Setting mode name
Setting
Returns to
1) Shifts pages when setting items at the same level
2) Moves to another item at the same level on the item
selection key
previous screen level.
the main screen.
item name
cannot be displayed on one page.
2.Name and Function of Each Part
1. Y-axis display section
・Measured value .....Displays Y-axis measured values such as pressure, and also
・Unit ....................... Displays the unit set with the analog CH set on the Y-axis.
2. High and low limit set values
・Displays the High Limit / Low Limit of the ch (operation ch) set by external input in the
Multi-Hold / Hysteresis mode.
・ Displays the HI-HI Limit / LO-LO Limit of the ch (operation ch) set by external input
in the Waveform Comparison / Waveform and Displacement Comparison mode.
(See "16-2-7. High Limit / Low Limit / HI-HI/LO-LO Mode / HI-HI Limit / LO-LO
Limit" P109.)
displays error messages when measurement errors occur.
(See "7-4. X/Y-axis Setting by Operation Mode and Input CH" P41,
and "19. Error Messages" P150.)
StatusDisplay color
NormalBlue
In samplingWhite
HoldGreen
(See "5-7. Unit" P35, and "16-2-3. Select Analog CH" P104.)
8
2.Name and Function of Each Part
WAIT S : Waiting for start.
WAIT 2 : Waiting for the rising edge of start/level.
WAIT L : Waiting for level.
SAMPLE : In sampling.
HIS GO : In hysteresis go sampling
HIS RT: In hysteresis return sampling.
COPY: In waveform copy
STOP: Waiting for start.
3. X-axis (displacement) display section
Displays X-axis measured values such as displacement in the Hysteresis / Hysteresis 2 /
Waveform and Displacement Comparison mode, and also displays error messages when
measurement errors occur.
(See "7-4. X/Y-axis Setting by Operation Mode and Input CH" P41, and "19. Error
Messages" P150)
4. Displacement high and low limit set values
Displays the displacement comparison values of the ch (operation ch) set by external
input in the Hysteresis / Hysteresis 2 / Waveform and Displacement Comparison mode.
Displacement comparison values that are displayed vary depending on the Displacement
High-Low Mode.
(See "16-2-8. Displacement High-low Mode / Displacement High Limit / Displacement Low
Limit / Displacement Hold High Limit / Displacement Hold Low Limit" P110.)
StatusDisplay Color
Thinning-outYellow
WA RN IN GR ed
ModeComparison value
Hold ValueDisplacement Hold High/Low Limit
Peak / BothDisplacement High/Low Limit
5. Lithium battery alarm
Displayed when the lithium battery requires replacement. (Normally, nothing is
displayed.)
If this alarm lights, replace the lithium battery in good time. (The life of the lithium
battery is approx. 7 years, which depends on the working environment.) For replacement
of the battery or handling of interchangeable batteries, contact us.
6. Status display section
ⅰ Status display
9
2.Name and Function of Each Part
WAIT: In tracking display
DETECT : In hold detection period.
HOLD: In data hold.
NOV: In NOV RAM writing.
HH: Indicated value > HI-HI Limit set value
HI: Indicated value > High Limit set value
OK: Neither of the High and Low Limits is exceeded.
LO: Indicated value < Low Limit set value
LL: Indicated value < LO-LO Limit set value)
Priority : NOV = HH = LL > HI = LO > OK
HI LO
HH LL
OK
or
HI LO
HH HL
OK
or
HOK
or
…
…
Hold judgment
Waveform
comparison
judgment
Displays the ch set by external input (operation ch).
ⅱ Hold status display
ⅲ High and low limits / NOV RAM writing display
Operation modeStatus display
Multi-Hold
Hysteresis
Hysteresis 2
Waveform Comparison
Waveform and
Displacement Comparison
ⅳ ch display
7. Set operation mode display
Displays the currently set Operation Mode.
8. Control signal display
Displays the T/H and H/M input status. (See "15-8. Control Signal Display" P99)
9. Display of high and low limit set values
Displays the High and Low Limit Set Values of the ch (operation ch) set by external input in the
Multi-Hold mode. (See "16-2-7. High Limit / Low Limit / HI-HI/LO-LO Mode / HI-HI Limit /
LO-LO Limit" P109), and displays the comparison waveform of the currently set opera-
tion ch in the Waveform Comparison / Waveform and Displacement Comparison mode.
(See "13-1. Function" P68.)
10
2.Name and Function of Each Part
1. AC power
2. Frame ground
3. RS-232C
connector
4. Calibration
5. Control connector
6. SI/F terminal
7. Analog
8. Optional slot
input connector
LOCK switch
input / output
terminals
2-2. Rear Panel
1. AC power input connector
Connect the attached AC power cord. The input voltage is 100 ~ 240V AC ( ± 10%), and
the frequency is 50/60Hz.
2. Frame ground (F.G)
This is a grounding terminal. Be sure to ground the FG terminal to prevent electric
shocks and failures caused by static electricity.
3. RS-232C connector
This is an RS-232C connector to send and receive measured data and status information.
The applicable plug is OMRON-manufactured XM2D-0901 (cover: XM2S-0913 <with
inch thread #4-40>) or an equivalent.
(See "4-2. RS-232C Connection" P24, and "18-2. RS-232C Interface" P133.)
11
4. Calibration LOCK switch
This is a LOCK switch to prevent calibrated values from being changed by mistake.
Changing of calibrated values is prohibited when this switch is ON. (See "5-5. LOCK
Release" P34.)
2.Name and Function of Each Part
5. Control connector
This is a connector to input and output control signals, operation ch settings, etc. The
applicable plug is DDK-manufactured 57-30500 (accessory) or an equivalent.
(See "4-3. Control Connector Connection" P25.)
6. SI/F terminal
This is a serial interface (SI/F) terminal block to connect a UNIPULSE-manufactured
printer, external display, data converter, etc.
(See "4-1-2. SI/F Connection" P23, and "18-1. SI/F 2-wire Serial Interface" P132)
7. Analog input / output terminals
These are terminals for strain gauge type sensor / displacement sensor inputs and voltage
outputs.
(See "4-1-1. Analog Input / Output Connection" P21.)
8. Optional slot
Up to four optional boards can be mounted.
12
3.Screen Configuration and Setting Methods
【Main】
【Mode Setting】
【Special】
【Numerical Value Input】
【Selection】
Each mode key
【Execution】
【Set Value Selection】
【Hold Point Display】
or
* Waveform
Waveform Sampling
Reading
Sampling
* Waveform
Reading
Waveform Editing
* Go to Select Rejected Waveform Reading from the selection screen.
Each setting key
Each
setting
key
3. Screen Configuration and Setting Methods
3-1. Screen Flow Chart
13
3.Screen Configuration and Setting Methods
【Waveform Reading 【Hold Point for
【Rejected Waveform 【Rejected Waveform
【Basic Setting】
[Code Setting]
Hold Setting key
【Judgment Setting】
【Hold Setting】
Category key
Waveform Reading】Graph】
Reading Selection】 Reading Graph】
Each
setting
key
【Hold Point for Rejected
Waveform Reading】
14
Mode set
PAGE1
CH1 CAL.
CH2 CAL.
Operation Mode
Code set
Wave sampling
Read Wave
Communication
Option set
System
System 2
CH2 CAL.
PAGE1
Axis(P41)
EXC. Volt(P34)
Zero CAL.(P35)
Equiv. CAL.(P36)
Actual CAL.(P37)
Digital Flt.(P38)
Analog Flt.(P38)
PAGE2
Decimal Place(P35)
Unit(P35)
Digital Offset(P38)
Increment(P35)
Read Wave
PAGE1
Read Wave(P92)
Read NG Wave(P96)
NG Wave CLR.(P97)
Control Signal (P99)
Y StartPoint(P97)
Y Scale(P97)
X StartPoint(P98)
X Scale(P98)
Communication
PAGE1
Speed(P133)
Data Length(P134)
Parity(P134)
Terminator(P134)
Communication Mode
(P134)
SI/F Hold Trans. (P134)
SI/F Hold Point(P134)
Main
Operation Mode
PAGE1
Operation Mode(P40)
Sample Rate(P40)
CH1 CAL.
PAGE1
EXC. Volt(P34)
Zero CAL.(P35)
Equiv. CAL.(P36)
Actual CAL.(P37)
Digital Flt.(P38)
Analog FIlt.(P38)
Decimal Place(P35)
PAGE2
Unit(P35)
Digital Offset(P38)
Increment(P35)
Wave Sampling
PAGE1
Wave Call(P82)
Clear(P82)
Sampling(P83)
Edit(P84)
3.Screen Configuration and Setting Methods
3-2. Setting Mode Tree Chart
15
3.Screen Configuration and Setting Methods
PAGE2
F-T Output(P120)
Auto Code Up(P107)
Code Up Point(P107)
[Basic]
PAGE1
Parameter Copy(P104)
Select CH(P104)
Wave Start Level (P113)
Wave End Level(P113)
Hold Timer ※(P105)
Inhibit Timer(P119)
Auto Code Up(P107)
Code Up Point(P107)
System
PAGE1
Protect(P127)
Display Rate(P127)
Backlight(P127)
Contrast(P128)
System 2
PAGE1
All Param. CLR.(P128)
Self Test(P128)
ROM-RAM CHK(P131)
NOVRAM CHK(P131)
PASS WORD(P131)
Code Set (Display items vary depending on the selected Operation Mode.)
HI-LO Relative(P108)
HI Limit(P109)
LO Limit(P109)
HH/LL Mode(P109)
HI-HI Limit(P109)
LO Limit(P109)
DPM HI-LO Mode (P110)
DPM HI(P110)
DPM LO(P110)
[Judgment]
PAGE1
HI-LO Relative(P108)
HI Limit(P109)
LO Limit(P109)
HH/LL Mode(P109)
HI-HI Limit(P109)
LO Limit(P109)
[Judgment]
PAGE1
HH/LL Mode(P109)
HI-HI Limit(P109)
LO Limit(P109)
U. Area Output(P80)
[Judgment]
PAGE1
HH/LL Mode(P109)
HI-HI Limit(P109)
LO Limit(P109)
DPM HI-LO Mode (P110)
DPM HI(P110)
DPM LO(P110)
DPM Hold HI(P110)
DPM Hold LO(P110)
U. Area Output(P80)
PAGE2
DPM Hold HI(P110)
DPM Hold LO(P110)
[Judgment]
PAGE1
HI-LO Relative(P108)
HI Limit(P109)
LO Limit(P109)
HH/LL Mode(P109)
HI-HI Limit(P109)
LO Limit(P109)
DPM HI-LO Mode (P110)
DPM HI(P110)
DPM LO(P110)
* Only the hold setting includes further setting items.
Setting items vary depending on the hold category
selection. For the items in the hold setting, see "16-2-4.
Hold Setting" P105.
Option Set
PAGE1
PUI
ODN
CCL
* For details, see the
attached Option Manual.
16
Loading...
+ 150 hidden pages
You need points to download manuals.
1 point = 1 manual.
You can buy points or you can get point for every manual you upload.