Unipulse F371 Operation Manual

F371
DIGITAL INDICATOR
20 Jul. 2011
Rev. 1.18

Cautions and Requests for Use

Do not disassemble the main body for modifications or repair.
Be sure to use crimp contacts for connection to terminal blocks, and do not to
connect bare wires as they are.
Be sure to ground the protective ground terminal.
- The main F.G. is indicated by , it must be grounded with the protected
earth.
Be sure to disconnect the power cable when performing the following.
- Attachment/detachment of connectors of options.
- Wiring/connection of cables to terminal blocks.
- Connection of the ground line.
Cautions and Requests for Use
Be sure to assort the machine which is supplied by AC power through the ter-
minal in a panel with lock or screws.
Carefully check wiring, etc. before applying power.
Take an interval of more than 5 seconds when repeating ON/OFF.
Take adequate shielding measures when using at the following locations.
- Near a power line.
- Where a strong electric field or magnetic field is formed.
- Where static electricity, relay noise or the like is generated.
Do not install in the following environments.
- Place exposed to direct sunlight.
- Where the temperature and/or humidity exceeds the range in the
specifications.
- Place containing corrosive gas or flammable gas.
- Place with large quantities of dust, salt or iron powder.
- Where the product may be splashed with water, oil or chemicals.
- Where the main body is directly affected by vibration or shock.
Set the correct Excitation Voltage for the sensor. (2.5V is set when F371 is dispatched from us.
Safety Precautions
WARNING
Misuse may cause the risk of death or serious
injury to persons.
Misuse may cause the risk of injury to persons or
damage to property.
CAUTION
Safety Precautions
Indications for safe use and their meanings
In this manual, precautions for using the F371 digital indicator safely are indicated as
follows. Be sure to follow the precautions given here because they are important
descriptions relating to safety.
Indications and their meanings are as follows:
Explanation of pictographs
The means a caution (or warning).
A specific description is written in the △ .
The illustration on the left-hand side shows "Caution: May explode".
The means a caution (or warning).
A specific description is written in the △ .
The illustration on the left-hand side shows a general caution.
About the built-in lithium battery
Never disassemble, deform under pressure or throw the battery
into fire.The battery may explode, catch fire or leak.
- Battery
Model: CR2477-1HF made by Matsushita Battery Industrial Co., Ltd.
Nominal voltage: 3V
Nominal electric capacity: 1000mAh
WARNING
For connection to the signal I/O terminal block, wire correctly after
checking the signal names and terminal block numbers.
Also, turn off the power of the main body before connection/wiring to the
signal I/O terminal block.
CAUTION
Safety Precautions
About the signal I/O terminal block

CONTENTS

CONTENTS
1APPEARANCE DESCRIPTION . . . . . . . . . . . . . . . . . . . . . . . . .1
1-1Front Panel . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . .1
1-1-1Touch Panel Display . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . 1
1-1-2Power Lamp . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . 1
1-2Rear Panel . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . .2
1-2-1AC Power Source Input Terminal Block . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . 4
1-2-2Frame Ground (Functional ground) . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . 4
1-2-3Optional Slot . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . 4
2SETTING MODES . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . .5
2-1F371 Screen Configuration . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . .5
2-2Setting Modes Tree . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . .6
2-3About a Setting Call . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . .8
3CONNECTION . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . .9
3-1Power Input Connection . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . .9
3-2Load Cell Connection . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . .10
3-3.SI/F Connection . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . .11
3-4.External I/O Connection . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . .12
3-4-1.External Output Connection . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . 12
3-4-2External Input Connection . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . 12
3-5RS-232C Interface Connection . . . . . . . . . . . . . . . . . . . . . . . . . . . .13
3-6.Voltage Output Connection . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . .14
3-6-1Connecting to Cage Clamp Terminal Block . . . . . . . . . . . . . . . . . . . . . . . . . . 15
4CALIBRATION . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . .16
4-1Equivalent Input Calibration Procedure . . . . . . . . . . . . . . . . . . . . . .17
4-2Actual Load Calibration Procedure . . . . . . . . . . . . . . . . . . . . . . . . .18
4-3Zero Calibration . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . .19
4-4Actual Load Calibration . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . .20
CONTENTS
4-5Equivalent Input Calibration . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . .21
4-6Digital Offset . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . .23
4-7Increment (This step may be omitted if there is no change.) . . . . .24
4-8Decimal Place . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . .25
4-9Calibration Protection . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . .26
4-10Calibration Value Selection . . . . . . . . . . . . . . . . . . . . . . . . . . . . . .27
4-10-1External Selection of Calibration Value . . . . . . . . . . . . . . . . . . . . . . . . . . . . 28
4-11.Unit . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . .29
5SETTING OF FUNCTIONS . . . . . . . . . . . . . . . . . . . . . . . . . . .31
5-1Digital Zero . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . .31
5-2Digital Filter . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . .32
5-3Analog Filter . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . .33
5-4Display Rate . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . .33
5-5Stability . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . .34
5-6Zero Track . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . .35
5-7Contrast Adjustment . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . .36
5-8.Backlight . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . .36
5-9Excitation Voltage . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . .36
5-10Auto Print . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . .37
5-11Hold Off Print . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . .39
5-12.Parameter Protection . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . .39
6COMPARISON FUNCTIONS . . . . . . . . . . . . . . . . . . . . . . . . . .40
6-1HI Limit/LO Limit/HI-HI Limit/LO-LO Limit . . . . . . . . . . . . . . . . . . . .41
6-2Hysteresis . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . .42
6-3Near Zero . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . .44
6-4HI-LO Limit Comparison Mode . . . . . . . . . . . . . . . . . . . . . . . . . . . .44
6-5HI-LO Limit Output Mode . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . .45
6-6Alarm Mode . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . .46
CONTENTS
7HOLD FUNCTIONS . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . .47
7-1Hold Setting --- common --- . . . . . . . . . . . . . . . . . . . . . . . . . . . . . .47
7-1-1Hold Mode . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . 47
7-1-2Hold Detection Time . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . 48
7-1-3Auto Start Level . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . 49
7-1-4Hold Point Shift . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . 49
7-2Hold Setting --- relative maximum and relative minimum --- . . . . .50
7-2-1Minimum Count . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . 51
7-2-2Valley Detection Level . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . 51
7-3Hold Setting --- inflection point --- . . . . . . . . . . . . . . . . . . . . . . . . .52
7-3-1Minimum Slope . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . 53
7-3-2Interval A and Interval B . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . 53
7-4Hold Setting --- mean value --- . . . . . . . . . . . . . . . . . . . . . . . . . . .55
7-4-1Mean Value Sampling Count . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . 55
7-5Hold Operation . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . .56
7-5-1Sample Hold . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . 56
7-5-2Peak Hold . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . 57
7-5-3Valley Hold . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . 58
7-5-4Peak-to-Peak (P-P) Hold . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . 59
7-5-5Relative Maximum and Relative Minimum Hold . . . . . . . . . . . . . . . . . . . . . . . 60
7-5-6Inflection Point Hold . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . 62
7-5-7Mean Value Hold . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . 63
7-6How to Specify the Hold Detection Period . . . . . . . . . . . . . . . . . . . .64
7-6-1All Period . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . 64
7-6-2Externally Specified Period Hold
(Peak, Valley , Peak-to-Peak and Mean Value) . . . . . . . . . . . . . . . . . . . . . . . 65
7-6-3Time Specified Period Hold
(Peak, Valley, Peak-to-Peak and Mean Value) . . . . . . . . . . . . . . . . . . . . . . . . 66
7-6-4Time Specified Period Hold with Trigger
(Peak, Valley, Peak-to-Peak and Mean Value) . . . . . . . . . . . . . . . . . . . . . . . . 67
8MULTI-HOLD FUNCTION . . . . . . . . . . . . . . . . . . . . . . . . . . . .68
8-1About Changing of the Setting CH . . . . . . . . . . . . . . . . . . . . . . . . . .69
9WAVEFORM DISPLAY . . . . . . . . . . . . . . . . . . . . . . . . . . . . . .70
9-1Graphic Display Screen . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . .70
9-1-1Hold Point Plotting . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . 71
9-1-2X-axis and Y-axis on the Graph Plotting Screen . . . . . . . . . . . . . . . . . . . . . . 71
CONTENTS
9-2Graph Plotting . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . .72
9-2-1Continuous . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . 72
9-2-2External . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . 73
9-2-3Level . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . 74
9-2-4External + Level . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . 74
9-3Graph Function . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . .75
9-4Interval Time . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . .75
9-5Graph Start Level . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . .75
9-6Level Detection Mode . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . .76
9-7X Start Point . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . .77
9-8Y (Load) Start Point and Y (Load) End Point . . . . . . . . . . . . . . . . . .78
10VOLTAGE OUTPUTVOL OUT) . . . . . . . . . . . . . . . . . . . . .79
11SCREEN LOCK . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . .80
12EVENT OUTPUT AT THE END OF GRAPH PLOTTING . . . .80
13RS-232C INTERFACE . . . . . . . . . . . . . . . . . . . . . . . . . . . . . .81
13-1Communication Specifications . . . . . . . . . . . . . . . . . . . . . . . . . . . .81
13-1-1Specifications . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . 81
13-1-2Connector Pin Assignment . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . 82
13-1-3Cable . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . 82
13-2RS-232C Interface Setting . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . .83
13-2-1Communication Mode . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . 83
13-2-2Baud Rate . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . 83
13-2-3Character Length . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . 83
13-2-4Parity Bit . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . 84
13-2-5Stop Bit . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . 84
13-2-6Terminator . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . 84
13-3Communication Modes . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . .85
13-4Communication Format . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . .85
CONTENTS
14BCD DATA OUTPUT(OPTION) . . . . . . . . . . . . . . . . . . . . . . .92
14-1Connector Pin Assignment . . . . . . . . . . . . . . . . . . . . . . . . . . . . . .92
14-2Logic Change . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . .93
14-3BCD Data Hold . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . .93
14-4Equivalent Circuit . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . .93
14-5Signal Timing . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . .94
14-6BCD Data Update Rate Selection . . . . . . . . . . . . . . . . . . . . . . . . .95
14-7.BCD Output Data Selection . . . . . . . . . . . . . . . . . . . . . . . . . . . . . .95
15RS-485 COMMUNICATION INTERFACE(OPTION) . . . . . . .96
15-1Communication Specifications . . . . . . . . . . . . . . . . . . . . . . . . . . . .96
15-2RS-485 connection . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . .96
15-3Communication Method . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . .99
15-4RS-485 Interface Setting . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . .100
15-4-1ID . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . 100
15-4-2Terminator Resistance . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . 100
15-4-3Communication Method . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . 100
16D/A CONVERTER(OPTION) . . . . . . . . . . . . . . . . . . . . . . . .101
16-1Voltage Zero and Full Scale, Current Zero and Full Scale . . . . .102
16-2D/A Output Mode . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . .103
16-3Voltage Output Selection . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . .103
16-4Current Output Selection . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . .103
17DC POWER SOURCE . . . . . . . . . . . . . . . . . . . . . . . . . . . . .105
18ERROR MESSAGES . . . . . . . . . . . . . . . . . . . . . . . . . . . . . .106
CONTENTS
19SELF-CHECK AND INITIALIZATION . . . . . . . . . . . . . . . . .107
19-1Self-Check . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . .107
19-1-1Self-Check DSP . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . 107
19-1-2Self-Check MEM . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . 107
19-1-3Self-Check KEY . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . 108
19-1-4Self-Check EXT . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . 108
19-2Initialization(All setting value clearances) . . . . . . . . . . . . . . . . . .109
19-3Password . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . .109
20ABOUT JAPANESE/ENGLISH DISPLAY SELECTION . . .110
21BLOCK DIAGRAM . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . .111
22DIMENSIONS . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . .112
23INSTALLATION IN A PANEL . . . . . . . . . . . . . . . . . . . . . . .113
24SPECIFICATIONS . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . .114
24-1Analog Section . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . .114
24-2Digital Section . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . .114
24-3Communication . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . .115
24-4Options . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . .115
24-5External Input and Output . . . . . . . . . . . . . . . . . . . . . . . . . . . . . .116
24-6General . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . .117
24-7Accessories . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . .118
25SETTING ITEM CHART . . . . . . . . . . . . . . . . . . . . . . . . . . . .119
26CONFORMITY TO EC DIRECTIVES . . . . . . . . . . . . . . . . . .123
CONTENTS

1. APPEARANCE DESCRIPTION

Touch Panel Display
Power Lamp
Hold Status Display
Display Change Button
Comparison Status
Indicated Value
Setting Input Screen Call Button
Digital Zero Button
Setting Mode Button
Display
Display
Comparison Display Hold Display Graph Display

1-1. Front Panel

1.APPEARANCE DESCRIPTION

1-1-1. Touch Panel Display

This is the touch panel display for displaying an indicated value and graph set value and
for setting various setting items of the F371. During measurement, a comparison display,
hold display and graph display can be selected according to the function in use.

1-1-2. Power Lamp

This lamp lights when the power to the F371 is on.
1
1.APPEARANCE DESCRIPTION
Frame ground
Signal I/O terminal
Optional slot
AC power source input terminal block
block
1GND
2 VOL OUT
3+SIG
4 -EXC
5-SIG
6+EXC
over
under
1
2
3
4
5
6
terminal block
terminal block
fix screw
fix screw
A1
COM1 B1
COM2
A2 OUT HH B2 IN CODE0
A3 OUT HI B3 IN CODE1
A4 OUT OK B4 IN CODE2
A5 OUT LO B5 IN CODE3
A6 OUT LL B6 IN T/H
A7
COM1 B7
COM2
A8 OUT H/E B8 IN H/M
A9 IN D/LOCK B9 IN D/Z
A10 OUT EVENT B10 IN ST/SP
A11 OUT SI/F B11 IN CAL0
A12 OUT SI/F B12 IN CAL1
A1 B1
A12 B1 2

1-2. Rear Panel

Terminal block
Signal input / output connect
2
1.APPEARANCE DESCRIPTION
COMPARISON
FUNCTIONS
page 40
The input for controlling the hold signal.
page 47HOLD FUNCTIONS
+EXC
-EXC +SIG
-SIG
See the section on Load Cell Connection page 10 for connection.
VOL OUT GND
See the section on Voltage Output Connection page 14 for connection.
COM1 The terminal common to output signals.
OK HI LO HH LL
H/E Outputs the hold end signal.
EVENT Outputs the event signal when graph plotting ends.
The terminal for connecting to a strain-gage sensor.
The voltage output terminal. page 79「VOLTAGE OUTPUT(VOL OUT)」
Outputs the OK signal.
Outputs the HI signal.
Outputs the LO signal.
Outputs the HH signal.
Outputs the LL signal.
             page 47HOLD FUNCTIONS
See the section on External I/O Connection page 12 for connection.
COM2 The terminal common to input signals.
CODE0 CODE1 CODE2 CODE3
T/H
H/M
D/Z The input for digital zero (making the indicated value zero).
ST/SP The graphic display start/stop signal.
CAL0 CAL1
D/LOCK Screen lock input for invalidating key operation on the measurement
See the section on External I/O Connection page 12 for connection.
Selects the CH No. for the multi-hold function. Calibration value selection input
          → page 68MULTI-HOLD FUNCTION
              → page 31Digital Zero
              → page 72Graph Plotting
Selects the CAL No. for calibration value selecttion function. page 28「External Selection of Calibration Value」
screen (comparison/hold/graph).
SI/F The 2-wire serial interface for coupling a UNIPULSE printer,
external display, etc.
See the section on SI/F Connection page 11 for connection.
3
1.APPEARANCE DESCRIPTION

1-2-1. AC Power Source Input Terminal Block

Connect the AC power cord. The input voltage is 100V - 240V AC, and the frequency is
50/60Hz.

1-2-2. Frame Ground (Functional ground)

This is a ground terminal block. Be sure to ground the F.G. terminal to prevent electric
shocks and failures due to static electricity.

1-2-3. Optional Slot

Any one of the following optional boards can be mounted.
• BCD data output F371
• D/A converter
• RS-485
• CC-Link interface (CC-Link is an abbreviation for "Control & Communication Link".)
• Device Net interface
4

2. SETTING MODES

HOME
MODE
UP
Mode keys
Setting keys
HOME
HOME
Set value entry from
the ordinary display screen
UP
[Ordinary display screen]
[Mode selection screen]
[Set value selection screen]
[Set value entry screen]
With the NEXT key on the set value entry screen, the next set value entry screen appears.

2-1. F371 Screen Configuration

2.SETTING MODES
5
2.SETTING MODES
ordinary display screen
Mode selection screen
OPERATION
PAGE1
Digital Filter (P32)
Analog Filter (P33)
Display Rate (P33)
Stability (Time) (P34)
Stability (Cnt) (P34)
PAGE2
Zero Track (Time) (P35)
Zero Track (Cnt) (P35)
Contrast1 (P36)
Contrast2 (P36)
Backlight (P36)
PAGE3
Ext Voltage (P36)
Auto Print (P37)
Hold Off Print (P39)
CAL. LOCK (P26)
PARA. LOCK (P39)
COMPARISON
PAGE1
HI_HI Limit (P41)
HI Limit (P41)
LO Limit (P41)
LO_LO Limit (P41)
Hysteresis (P42)
PAGE2
Near Zero (P44)
Comparison Func (P44)
Comparison Mode (P45)
Alarm Mode (P46)
HOLD
PAGE1
Hold Function (P47)
Hold Detect Time (P48)
Auto Start Level (P49)
Minimum Count (P51)
Valley Detect Lv1 (P51)
PAGE2
Minimum Slope (P53)
Interval A (P53)
Interval B (P53)
Hold Point Shift (P49)
Mean Value Sampling Count (P55)
GRAPH
PAGE1
Graph Function (P75)
Interval Time (P75)
Graph Start Level (P75)
Level Detect Mode (P76)
X Start Point (P77)
PAGE2
Y Start Point (P78)
Y End Point (P78)

2-2. Setting Modes Tree

6
2.SETTING MODES
CALIBRATION
PAGE1
Zero Cal. (P19)
Actual Cal. (P20)
Equiv. Cal. (P21)
CAL. Select (P27)
Increment (P24)
PAGE2
Unit (P29)
Decimal Place (P25)
Digital Offset (P23)
OPTION
A setting function changes
by the option.
SYSTEM
PAGE1
All Para. Clear (P109)
SelfCheck DSP (P107)
SelfCheck MEM (P107)
SelfCheck KEY (P108)
SelfCheck EXT (P108)
PAGE2
PASSWORD (P109)
LANGUAGE (P110)
BCD OUTPUT
PAGE1
BCD Data Update Rate (P95)
BCD Output Data Selection
(P95)
RS-232C
PAGE1
Communication Modes (P83)
Baud Rate (P83)
Character Length (P83)
Parity Bit (P84)
Stop Bit (P84)
PAGE2
Terminator (P84)
D/AOutput
PAGE1
D/A Output Mode (P103)
D/A Zero Output (P102)
D/A Zero and Full Scale
(P102)
Current Zero Output
(P102)
Current Zero and Full
(P102)
RS-485
PAGE1
Communication Modes(P83
)
Baud Rate (P83)
Character Length (P83)
Parity Bit (P84)
Stop Bits (P84)
PAGE2
Terminator (P84)
ID
(P100)
Terminator resistance
PAGE2
Voltage Output Selection (P103)
Current Output Selection (P103)
7
2.SETTING MODES
Page 3
↑↑
Mode Page
→→
Setting call
Operation setting
Setting call
Can be changed by the PAGE
Operation setting Comparison settingHold Communication settingCalibration Graph settingSystem Option
Modes are as follows:
Operation setting
button.
PAG E
PAGE
Page 1 Page 2 Page 3
PAG E

2-3. About a Setting Call

In this manual, a setting function call is described as follows.
Calibration Protection
This call can be made by the following procedure.
1Press the MODE button on the ordinary display screen.
2The mode setting screen appears. Select the mode.
3The setting function setting screen appears. Select the function.
8

3. CONNECTION

Black
White)
Green)
AC
IN
L
N
F. G.
CAUTION
Pay close attention to the wring in case of power on.
Red
Black)
DC IN
+ -
F. G.
CAUTION
Be aware that the voltage drops depending on the wire thickness
and length. Also, never input an AC power source. Doing so will
cause a failure.

3-1. Power Input Connection

AC spec.
Connect AC power code. The input voltage is 100V-240V AC.
3.CONNECTION
The frequency is 50/60Hz.
DC spec.(Depending on the request at the time of order)
Connect the positive (+) side of the power source to the red screw side of the terminal
block on the back of the F371, and its negative (-) side to the black screw side. The input
voltage is 12V-24V DC.
9
3.CONNECTION
+EXC
-EXC
+SIG
-SIG
F.G or
IN
OUT
IN
OUT
A1
B2
A2
B1
F371
+EXC
-EXC
+SIG
-SIG
+IN
-OUT
-IN
+OUT
A1
B2
A2
B1
F371
F.G or

3-2. Load Cell Connection

4-wire sensor
6-wire sensor
Short-circuit +EXC with +S and -EXC with -S for connecting a 6-wire strain-gage
sensor.
10

3-3. SI/F Connection

Converter
Display
Printer
SI/F
B10
B11
Inside Outside
F371
The 2-wire serial interface has connective ability for coupling a UNIPULSE printer,
external display, etc.
The interface is nonpolarized and up to three external instruments may be connected.
A two-core parallel cable or a cabtyre cable may be used for connection.
3.CONNECTION
11
3.CONNECTION
COM1
Spark Killer
Load
F371 Inside
Spark Killer
DC
Var isto r
Load
AC
Relay
Vex t
A3
Supply an external power source for a relay drive
Output Data  Tr
0  OFF
1   ON
Output Transistor Status
power source (Vext).
Power Supply
Power Supply
COM2
Push Switch
Toggle Switch
Relay
+12V
TTL Open Collector (ON when IN is HI)
F371 Inside
A6
IN
IN
Open OFF Short ON
- The external element is required to withstand Ic=10mA.
- Leakage from the external element is required to be 100μA or below.

3-4. External I/O Connection

3-4-1. External Output Connection

The external output circuit is operated through an open collector. A3 COM1 is the
common terminal. The open collector output capacity is 30mA and the withstand voltage
is up to 30V.
Equivalent circuit

3-4-2. External Input Connection

A signal is inputted to the signal input circuit by short-circuiting or opening the input
terminal and the COM2 terminal. Short-circuiting is effected by means of a contact (such
as a relay or a switch) or a noncontact (such as a transistor or an open-collector TTL).
12

3-5. RS-232C Interface Connection

Insert
6
7
8
5
4
2
1
3
Metal case
Fit the metal cases into the Din connector,
and cap with the plastic covers.
Plastic cover
Wire connection surface
Side view
Secure with hardware
Case
F.
TxD
RxD
DTR
GND
GND
DTR
TxD
RxD
CD
DSR
CTS
RTS
F. G
TCP8080-01-520
F371
D-sub 9 pin
Personal computer
, etc.
Cabling diagram
CA371-232 (optional)
This connector connects the RS-232C.
Pin No. Signal name
1
2TxD
3RxD
4DTR
5
6
7GND
8
Case F.G.
3.CONNECTION
Example of cabling
The following shows an example of connection between DTE-DTE terminals. This will
require modification depending on the equipment to be connected. For details, see the
instruction manual of the equipment to be connected.
13
3.CONNECTION
+
-
F.G o r
External
2
1
SIG
COM
VOL OUT
GND
F371 inside outside
equipment

3-6. Voltage Output Connection

The voltage output terminal is an interface to extract analog voltage proportional to
sensor signal inputs.
Since the VOL OUT terminal is not insulated from the internal circuit, use a shielded
cable for connection with the external equipment and wire within 2 - 3m.
Do not short-circuit. Doing so will cause a failure.
Do not apply voltage from the outside. Doing so will cause breakage.
14

3-6-1. Connecting to Cage Clamp Terminal Block

5 - 6mm
CAUTION
Cable can be from 24 to 14AWG (0.2 to 2.5mm2)
It is not necessary to solder the cable wires or to fix a solderless terminal.
If several cables to be inserted to the same hole, twist those cable wires
together and insert.
If you connect a cable (a load cell, SI/F, external input and output),
please turn off and be sure to perform the power supply of a main part.
The output terminal D/A option and RS-485 option is using the cage clamp system
terminal stand. Please connect in the following procedure.
1.Strip the casing 0.2inch (6mm) on the cable to be connected.
2.Twist the bare wire to fit the terminal hole.
3.Insert the supplied screwdriver into the upper hole and lift upward.
3.CONNECTION
4.Insert the twisted wires into the lower hole.
5.Make sure cable is clamped securely and does not come out with a slight tug.
15

4.CALIBRATION

F371
Strain gauge sensor
Rated output value (mV/V)
Indicated value
A data sheet is attached to a strain-gage sensor at the time of purchase.
The data sheet provides data including:
capacity load (in kg, t, etc.)
rated output voltage (in mV/V)
non-linearity, hysteresis, input resistance, output resistance
and zero balance.
Enter the capacity and the rated output value required for equivalent
input calibration into the F371.
4. CALIBRATION
Calibration is performed for matching the F371 to a strain-gage sensor. The following
two types of calibration are available for the F371.
Equivalent Input Calibration
Calibration is performed without an actual load by entering the rated output value (mV/
V) and the capacity (to be indicated) of the strain-gage sensor by the keys. Calibration is
easily performed when no actual load is available.
For example, the gain is automatically determined by entering:
2.001mV/V (rated output) - 100.0kg (capacity)
as indicated for a load.
16
Actual Load Calibration
Actual
Strain gauge sensor
Indicated value
F371
load
Protection Release
Increment
Input Calibration
Calibration Protection
Decimal Place
Digital Offset
4-9.
4-7.
4-8.
4-5.
4-6.
4-9.
Zero Calibration
4-3.
Value Selection
4-10.
Calibration
Equivalent
Calibration
Apply an actual load to the strain-gage sensor and enter the actual load value by the keys
for calibration. Calibration is accurately performed with reductions in errors.

4-1. Equivalent Input Calibration Procedure

Follow the steps below to perform equivalent input calibration.
4.CALIBRATION
Release the calibration protection.
Set the calibration value No.
(This step may be omitted for use with only one
selection.)
Enter the minimum value of digital increments.
(This step may be omitted if there is no change.)
Determine the decimal point place of the indicated value.
(This step may be omitted if there is no change.)
Set the zero point of the strain-gage sensor in no-load
condition (with the sensor unloaded).
Enter the rated output value and reading of the strain-gage
sensor.
The calibrated value can be offset in advance.
(This step may be omitted if there is no change.)
Turn on the calibration protection for preventing
misoperation.
17
4.CALIBRATION
Protection Release
Increment
Load Calibration
Calibration Protection
Decimal Place
Digital Offset
4-9.
4-7.
4-8.
4-4.
4-6.
4-9.
Zero Calibration
4-3.
Value Selection
4-10.
Calibration
Actual
Calibration

4-2. Actual Load Calibration Procedure

Follow the steps below to perform actual load calibration.
Release the calibration protection.
Set the calibration value No.
(This step may be omitted for use with only one
selection.)
Enter the minimum value of digital increments.
(This step may be omitted if there is no change.)
Determine the decimal point place of the indicated value.
(This step may be omitted if there is no change.)
Enter the zero point of the strain-gage sensor with no load
applied to the sensor.
Enter the span (gain) point of the strain-gage sensor with a
load applied to the sensor.
The calibrated value can be offset in advance.
(This step may be omitted if there is no change.)
Turn on the calibration protection for preventing
misoperation.
18

4-3. Zero Calibration

How to set
Setting call Page 1
→→
Calibration
Set the zero point in no-load condition.
1Press the MODE button.
4.CALIBRATION
2Press the CALIBRATION button.
3Press the Zero Cal.button.
4Press OK button after confirming no-load was applied to the sensor.
19
4.CALIBRATION
How to set
Setting call Page 1
→→
Calibration

4-4. Actual Load Calibration

Set the actual load value under an actual load.
1Press the MODE button.
2Press the CALIBRATION button.
3Press the Actual Cal. button.
4Apply an actual load to the sensor, enter the actual load value by the numerical keys
and determine with the OK button.
20

4-5. Equivalent Input Calibration

How to set
Setting call Page 1
→→
Calibration
Set the rated output value and reading of the sensor.
Rated output value: 0.000 - 3.000mV/V
Rated value: 00000 - 99999
1Press the MODE button.
4.CALIBRATION
2Press the CALIBRATION button.
3Press the Equiv.Cal. button.
21
4.CALIBRATION
4)Enter the rated output of the sensor by the numerical keys and determine with the OK
button. The decimal point is fixed.
5Enter the rated value by the numerical keys and determine with the OK button.
22

4-6. Digital Offset

How to set
99999 - 99999
Setting range:
Setting call Page 2
→→
Calibration
By using the digital offset function, the value obtained by subtracting the set value from
the indicated value is displayed.
This function is convenient when zero cannot be obtained with no load for some reason
or for offsetting.
(displayed value)=(actual indicated value)-(offset value)
1Press the MODE button.
4.CALIBRATION
2Press the CALIBRATION button.
3Press the PAGE button to select the page and press the Digital Offset button.
4)Enter the digital offset value by the numerical keys and determine with the OK button.
23
4.CALIBRATION
How to set
Setting call Page 1
→→
Calibration

4-7. Increment (This step may be omitted if there is no change.)

Set the increment of the indicated value.
Input range: 001 - 100
1Press the MODE button.
2Press the CALIBRATION button.
3Press the INCREMENT button.
4Enter the increment by the numerical keys and determine with the OK button.
24

4-8. Decimal Place

How to set
Setting call Page 2
→→
Calibration
Set the decimal point place of the indicated value. Selection can be made from
the following.
1Press the MODE button.
4.CALIBRATION
0, 0.0, 0.00, 0.000, 0.0000
2Press the CALIBRATION button.
3Press the PAGE button to select the page and press the DECIMAL PLACE button.
4Select the decimal place and determine with the OK button.
25
4.CALIBRATION
How to set
Setting call Page 3
→→
Calibration

4-9. Calibration Protection

Calibration-related set values can be protected so that they will not be changed
by misoperation. When Cal. Protect is ON, no change can be made while the
alarm sounds.
ON: Protected
OFF: Unprotected
1Press the MODE button.
2Press the CALIBRATION button.
3Press the PAGE button to select the page twice and Press the CAL Lock button.
4Select the decimal place and determine with the OK button.
26

4-10. Calibration Value Selection

How to set
Setting call Page 1
→→
Calibration
By storing up to four calibration values in the memory, the desired calibration value can
be called to switch the indicated value. Setting values that can be switched are as
follows:
Calibration Mode Setting Operation Mode Setting
Zero Calibration Excitation Voltage
Actual Load Calibration
Equivalent Input Calibration
Minimum Slope
Unit
Decimal Place
Digital Offset
4.CALIBRATION
When “Ext Input” is selected, the calibration value can be selected through external
selector signal CAL0 and CAL1.
1Press the MODE button.
2Press the CALIBRATION button.
27
4.CALIBRATION
B7 COM2
B11 CAL0
B12 CAL1
NOTE
It takes one second at maximum for
the changed calibration value to
become effective. During this time,
the calibration value is indefinable.
Also, the indicated value is
accordingly indefinable.
I/O connector
3Press the CAL. SELECT button.
4Select the calibration value from 0 - 3 and determine with the OK button.

4-10-1. External Selection of Calibration Value

With this function, four types of calibration values can be selected with external selector
signals CAL0 and CAL1 (when the calibration value selection setting is external).
Normally, when there is no input to CAL0 and CAL1 (the terminals are open),
calibration value 0 is selected. When each terminal is in the following condition, each
calibration value is selected.
CAL1 CAL0 Calibration value
Open Open alibration value 0
Open Short-circuit alibration value 1
Short-circuit Open alibration value 2
Short-circuit Short-circuit alibration value 3
28

4-11. Unit

4.CALIBRATION
Set the unit of the load to be displayed. It can be selected from the following.
Category Unit
Mass
Force
Pressure
Density
Momentum
Viscosity
Length
Speed
Flow rate
kgMggmg,μgtlb dynekdyneozTONNE,%
NkNMNmN,μNNmkNm MNmmNm,μNm ftlbinlbinozkgmgcm
PakPaMPaGPahPamPa,μPa barmbar,μbarmmHg,N/m
ftH
2OinH2Opsiapsigatm
3
g/cm3,t/m3,g/mlg/lkg/mmg/m
kg/m
kgm/skgm
2
/skgm
2
2
PaSmPaS
kmmcmmm,μm
m/skm/hm/s
2
rpm
kg/hkg/minkg/st/ht/st/min
3
m
/h,m3/min,m3/s,l/h,l/minl/s
29
4.CALIBRATION
How to set
Setting call Page 2
→→
Calibration
Scroll with the and buttons.
Scroll with the and buttons.
* A change of unit does not affect the indicated value (calibration value)
.
1Press the MODE button.
2Press the CALIBRATION button.
3Press the PAGE button to select the page and press the UNIT button.
4Select the category and determine with the OK button.
5Select the unit and determine with the OK button.
30

5. SETTING OF FUNCTIONS

OK
Cancel

5-1. Digital Zero

The indicated values is forcedly zeroed.
Digital Zero by means of Keys
1Press the DZ button.
5.SETTING OF FUNCTIONS
2)Press ‘OK’ to perform the digital zero. Press ‘Cancel’ to cancel the digital zero.
31
5.SETTING OF FUNCTIONS
How to set
D/Z
OFF
ON
Keep on for 1.0 msec. or more.
Note
Digital zero is reseted in case of power failure. Please set digital zero
again.
Setting call Page 1
Operation Setting
* See the section on About a
Setting Call page 8 .
Scroll button
Digital Zero by means of Keys
The digital zero may be performed by short circuiting the D/Z to the COM2 on the rear
panel signal I/O terminal block.

5-2. Digital Filter

The digital filter is a function for reducing drifts of the indicated value by means of a
moving average of data converted from analog to digital. The number of the moving
averages can be selected a range between 0 and 512.
With an increase in the number of filterings, the indicated value becomes more stable,
but the response to inputs becomes slower.
The value of a digital filter is scrolled with a button.
Number of settings:OFF, 2, 4, 8, 16, 32, 64, 128, 256, 512
32

5-3. Analog Filter

How to set
How to set
Setting call Page 1
Operation Setting
Setting call Page 1
Operation Setting
A low-pass filter is provided for filtering input signals from the strain-gage sensor and
canceling noise components.
The cut-off frequency can be selected in a range between 10Hz and 300Hz. With an
increase in the cut-off frequency, the response becomes faster, but noise components
may be indicated.
5.SETTING OF FUNCTIONS
Cut-off frequency: 10Hz, 30Hz, 100Hz, 300Hz

5-4. Display Rate

Enter the rate of rewriting the display.
The display rate can be selected in a range between 1 and 10 times/sec. The internal
operation speed does not change.
33
5.SETTING OF FUNCTIONS
How to set
Set
Indicated Value
50msec
Time
Set Period
range
Setting call Page 1
Operation Setting
CAUTION
When the time is 0.0 sec. and the width is 00 markings, stability is not
detected. At this time, the stable-time digital filter is normally off.
Turning on/off the near zero function is closely related to the auto print
function and HI-LO limit comparison mode.
For details, see the sections on Auto Print page 37 and HI-LO Limit
Comparison Mode page 44 .

5-5. Stability

Enter the parameters to detect stability.
If the difference between the current indicated value and the 50-msec-previously
indicated value is less than the set count and the duration of the condition is more than
the set time, the indicated value is regarded to be stable.
When stability is detected, a digital filter (fixed 32 times) to control instability in weight
value is automatically inserted. This stable-time digital filter differs from the digital filter
setting in the operation mode.
Setting range
- MD (time): 0.1 - 9.9 sec.
- MD (width): 01 - 99
34

5-6. Zero Track

How to set
When displacement of the zero point is within the set count of tracking
and it continues more than the set time, it is automatically made zero by
Zero Tracking function.
The time (tracking delay) is set in the range of 0.1 - 9.9 sec., and the
band (tracking band) is set in the range of 01 to 99.
If the time is set at 0.0 sec. and the band at 00, the zero tracking function
does not work.
Boundary of zero track
DELAY
Indicated value
+COUNT
-COUNT
BAND
0
DELAY
Band=count × 2
From the point when it returned within the range, counting will be resumed.
CAUTION
Since the zero track function works from the calibrated zero point of the
indicated value, it does not work if the indicated value is beyond the
track band. In such a case, obtain the zero point again by zero
calibration.
Setting call Page 1
Operation Setting
Gradual changes in the zero point due to drifts etc., are automatically tracked for
correction.
5.SETTING OF FUNCTIONS
Setting range
ZT (time): 0.0 - 9.9 sec.
ZT (width): 00 - 99
35
5.SETTING OF FUNCTIONS
How to set
How to set
How to set
Setting call Page 2
Operation Setting
Setting call Page 2
Operation Setting
Setting call Page 3
Operation Setting

5-7. Contrast Adjustment

Adjust the contrast of the touch panel display.
Brightness can be adjusted by CONTRAST1.
Screen flickering can be adjusted by CONTRAST2.

5-8. Backlight

The backlight is turned off if no button operation is performed for the set time (minutes).
The backlight is turned on by touching the panel. Setting 00 disables this function.
Setting range: 00 - 99 min.

5-9. Excitation Voltage

Select the bridge voltage to be supplied to the strain-gage sensor.
The bridge voltage is selectable from 10V, 5V and 2.5V.
Be sure to perform calibration after changing this setting.
36

5-10. Auto Print

How to set
Setting call Page 3
→→
Calibration
Indicated Value
Near Zero
0
Stable
Near Zero
ON
ON
ON
ON
OFF
OFF
OFF
Autmatic
Hold
1.5Seconds
Time
ON
Hold function……
Indicatid ValueSI/FCompariosn
Sensor input value
 Output signal
Printing
The indicated value is automatically printed to the UNIPULSE printer coupled with the
F371 through the SI/F. A print is made when the indicated value is stable. (Set the
stability parameters by the stability operation.)
When Near Zero is OFF, Indicated value is held until Near Zero turns ON after Stability
turned ON.
(Hold will be released when 1.5 sec. is passed after Near Zero is ON.)
Operation of the indicated value hold function
5.SETTING OF FUNCTIONS
37
5.SETTING OF FUNCTIONS
Near Zero
0
Time
Stable
Near Zero
ON
ON
ON
ON
OFF
OFF
OFF
Autmatic
Hold
1.5Seconds
ON
Indicated
Sensor input value
Printing
Va l ue
Auto Print is not performed under the following conditions.
Motion Detect
 Time: 0.0 sec. Width: 0.0 of divisions is set.
Print while hold is released.
 When Print is set.
Hold Mode
 When any items except Tracking is selected.
38

5-11. Hold Off Print

How to set
How to set
Setting call Page 3
→→
Calibration
Setting call Page 3
→→
Calibration
For the parameters which are protected by parameter protection,
see the section on SETTING ITEM CHART page 119.
At hold-off time, the held value is automatically printed to the UNIPULSE printer
coupled with the F371 through the SI/F.
(The hold is released by the T/H signal off timing when the period setting is all interval
in various hold modes by hold functions, and by the T/H signal of timing when other
periods are set.)

5-12. Parameter Protection

5.SETTING OF FUNCTIONS
Parameters are protected from being changed by misoperation.
39

6.COMPARISON FUNCTIONS

Time
Indicated value
HI Limit
LO Limit
HI
LO
OFF
ON
OFF
ON
HI-HI Limit
LL Limit
OFF
ON
OFF
ON
HH
LL
OFF
ON
OK
Va lu e
Va lu e
Va lu e
Va lu e
6. COMPARISON FUNCTIONS
By the comparison function, the HI limit and LO limit values are set, and when the
indicated value exceeds the HI limit, the HI output is turned on, and when the indicated
value falls below the LO limit, the LO output is turned on. Also, HI-HI limit and LO-LO
limit values may be set outside the HI-LO limit comparison. When the indicated value
exceeds the HI-HI limit, the HH output is turned on, and when the indicated value falls
below the LO-LO limit, the LL output is turned on. When the HI, HH, LO and LL
outputs are all off, the OK output is turned on.
<HI/LO output conditions>
HI: indicated value > HI limit value
LO: indicated value < LO limit value
<HH/LL output conditions>
HH: indicated value > HI-HI limit value
LL: indicated value < LO-LO limit value
<OK output conditions>
OK: All conditions of HH, HI, LO and LL are off.
40

6-1. HI Limit/LO Limit/HI-HI Limit/LO-LO Limit

How to set
Setting call Page 1
→→
Comparison setting
Press any of the HI-HI, HI, LO and LO-LO buttons at the bottom of the indicated
value display screen to go direct to the entry screen.
Simple setting call
6.COMPARISON FUNCTIONS
About the indicated value display color
The display color can be changed by pressing the indicated value display section. Every
time it is pressed, State 1 and State 2 are changed.
State 1
The indicated value display color is fixed (yellow).
State 2
The indicated value display color changes following the comparison status.
OK: green
HI, LO: yellow
HH, LL: red
41
6.COMPARISON FUNCTIONS

6-2. Hysteresis

The hysteresis value may be determined so as to allow a margin for timing the turning off
of the HI-LO limit comparison. Normally, it is turned on when the indicated value
exceeds the HI limit and is turned off when the indicated value falls below it. However,
by setting the hysteresis, it is turned off when the indicated value falls below the HI limit
further lowered by the hysteresis value.
This function is effective to prevent chattering in such a case where signals fluctuate
(vibrate) subtly.
Comparison conditions
- HI limit
ON condition: indicated value>HI limit value
OFF condition: indicated value<=(HI limit value-hysteresis value)
- LO limit
ON condition: indicated value<LO limit value
OFF condition: indicated value>=(LO limit value+hysteresis value)
- HI-HI limit
ON condition: indicated value>HI-HI limit value
OFF condition: indicated value<=(HI-HI limit value-hysteresis value)
- LO-LO limit
ON condition: indicated value<LO-LO limit value
OFF condition: indicated value>=(LO-LO limit value+hysteresis value)
42
Hysteresis operation
How to set
HI Limit
LO Limit
0
Hysteresis Range
Time
HI Output HI Status Display
LO Output LO Status Display
OK Status Display
Output ON Status ON
Output OFF Status OFF
Status ON Status ON
Output ON Status ON
Output ON Status ON
Output OFF Status OFF
Output OFF Status OFF
Indicated Value
OK Output
(Exsample : HI Output and LO Output and OK output)
Value
Value
Setting call Page 1
→→
Comparison setting
Hysteresis setting value is common to all HI limit value.
6.COMPARISON FUNCTIONS
43
6.COMPARISON FUNCTIONS
How to set
How to set
Turning on/off the near zero function is closely related to the auto print
function and HI-LO limit comparison.
For details, see the sections on HI-LO Limit Comparison Mode page 44
and Auto Print page 37 .
Setting call Page 2
→→
Comparison setting
Setting call Page 2
→→
Comparison setting

6-3. Near Zero

By this function, it is detected that the indicated value is near zero.
Near-zero ON:| indicated value | <=near zero set value
Near-zero OFF:| indicated value | > near zero set value
Setting range: 00000 - 99999

6-4. HI-LO Limit Comparison Mode

Set the operating condition of HI-LO limit comparison. Select the condition from the
following.
Continuous HI-LO limit comparison is performed continuously.
MD :HI-LO limit comparison is performed when the indicated value is
stable. Set the stability parameters by the stability operation.
NZ :HI-LO limit comparison is performed when the indicated value is
not near zero. Set the near zero parameters by the near zero operation.
MD+NZ :HI-LO limit comparison is performed when the indicated value is
stable and not near zero.
44

6-5. HI-LO Limit Output Mode

How to set
The HI limit output is turned on when the indicated value becomes
larger than the set value.
The LO limit output is turned on when the indicated value becomes
smaller than the set value.
COUTION
Even if any mode other than MODE 2 is selected, the name of
each setting does not change. Only the operation differs.
Setting call Page 2
→→
Comparison setting
The number of HI-LO limits can be changed..
Mode HI Limit Operation LO Limit Operation
6.COMPARISON FUNCTIONS
Mode 0
Mode 1
Mode 2
Mode 3 HI-HI Limit
Mode 4 None
HI-HI LimitHI Limit LO LimitLO-LO Limit
HI-HI LimitHI Limit LO Limit
HI-HI LimitHI Limit LO LimitLO-LO Limit
None
LO-LO Limit
HI LimitLO Limit LO-LO Limit
HI-HI LimitHI Limit LO LimitLO-LO Limit
45
6.COMPARISON FUNCTIONS
How to set
Hold values
HI-HI Limit (HH)
HI Limit (HI)
Real-time values
HH
HI
OFF
ON
OFF
HH
HI
OFF
Alarm mode unavailable
Alarm mode available
When the comparison output mode is “mode 0: HI limit operation for
all of HI-HI limit/HI limit/LO limit/ LO-LO limit”, the alarm mode
operation can only be performed with the HI-HI limit.
In a like manner, when the mode is “mode 4: LO limit operation for all
of HI-HI limit/HI limit/LO limit/LO-LO limit”, the alarm mode operation
can only be performed with the low-low limit.
Setting call Page 2
→→
Comparison setting

6-6. Alarm Mode

The comparison target of the HI-HI limit and LO-LO limit set values can be changed
from “hold value” to “real-time value”. By this function, whether or not the indicated
value becomes abnormal during hold can be monitored.
Unavailable : Comparison operation is performed with hold values.
Available : Comparison operation is performed with real-time values.
46

7. HOLD FUNCTIONS

How to set
Setting call Page 1
Hold
By the hold function including sample hold, peak hold, valley hold, peak-to-peak hold,
relative maximum and minimum hold and inflection point hold, a specific point in a
waveform is taken out for HI-LO limit comparison.
The operation of each hold will be described in detail.

7-1. Hold Setting --- common ---

7-1-1. Hold Mode

The F371 includes nine hold modes as shown in the table below.
In the peak, valley, peak-to-peak and mean value modes, period setting is required.
7.HOLD FUNCTIONS
Select all period, external signal, time, or time with trigger.
If you do not use any hold function, be sure to set the hold mode to tracking. (In tracking
condition, hold operation is not performed but input values are always displayed.)
HOLD MODE HOLD PERIOD SETTING
Tracking None
Sample None
Peak
Valley
P-P
Relative maximum
Relative minimum
Inflection Point
Mean value Required
Required
None
There is a section setup.
All Period External Signal Time Time with trigger
47
7.HOLD FUNCTIONS
How to set
Press the MODE button at the bottom of the indicated value display screen
to go direct to the hold mode entry screen.
Simple setting call
Setting call Page 1
Hold
Press the TIME button at the bottom of the indicated value display screen
to go direct to the hold detection time entry screen.
Simple setting call

7-1-2. Hold Detection Time

If you set the hold period setting in the hold function setting to Time or Time with
Trigger, set the time.
Setting range: 0.001 - 9.999 sec.
48

7-1-3. Auto Start Level

How to set
How to set
Setting call Page 1
Hold
In each hold mode of relative maximum/relative minimum/inflection
point, if the auto start level is set at 99999, the hold operation starts
only with the period signal H/M without performing level detection.
The actual hold point
t
w
When the hold point shift is set at 8, the data eight pieces before the actual hold point is held.
Caution
This does not work for other holds.
The retroactive sampling data is held irrespective of the hold
period. If the set numerical value is too large, point(s) out of the
period may be held.
Setting call Page 2
Hold
If you select the time specified period mode with trigger, or relative maximum, relative
minimum or inflection point hold in the hold mode setting, set the start level.
Setting range: -99999 - 99999
7.HOLD FUNCTIONS

7-1-4. Hold Point Shift

In the “sample hold” and “inflection point hold”, the sampling data is held retroactively
by the numerical value set under hold point shift.
Setting range: -99999 - 99999
49
7.HOLD FUNCTIONS
A
B
O
P
Q
2X
X
1.5X 3X
True relative maximum
True relative minimum
Relative
A
X
B
minimum
Relative maximum

7-2. Hold Setting --- relative maximum and relative minimum ---

If you select the relative maximum and relative minimum hold in the hold function setting, set the relative maximum and relative minimum value detection parameters "minimum count" and "valley detection level". Set referring to the principle of operation only when the value cannot be held successfully with the factory settings or when further adjustments are required.
Detection of the relative maximum value and relative minimum value
The logic of detecting the relative maximum value and relative minimum value is given
below.
First, when difference X between point
A and point B is larger than the
minimum count, point A is judged to be
the relative maximum value and point B
is judged to be the relative minimum
value.
When difference X between detected
relative maximum value A and relative
minimum value B exceeds the
predetermined detection levels (1.5, 2,
and 3 times), A is displayed in the rela-
tive maximum value hold mode at
respective points O, P and Q and the
value is held.
If the minimum count is too small, when
50
the waveform includes noise as shown
on the left-hand side, the noise is
regarded as the relative maximum value
or relative minimum value and a correct
value may not be held. In such a case,
set the minimum count somewhat large.

7-2-1. Minimum Count

How to set
How to set
Setting call Page 1
Hold
Setting call Page 1
Hold
Setting range: 0001 - 9999

7-2-2. Valley Detection Level

Level: 1/4, 1/2, 3/4, 1, 1.25, 1.5, 2, 3 times
7.HOLD FUNCTIONS
51
7.HOLD FUNCTIONS
D
C
C
AB
a
Input Signal
A : IntervalA B : IntervalB D : The amount of change
Inflection point Held as an inflection point if D > inflection judgment value.

7-3. Hold Setting --- inflection point ---

If you select the inflection point hold in the hold function setting, set the inflection point
detection parameters "minimum slope", "interval A" and "interval B".
Set referring to the principle of operation only when the value cannot be held
successfully with the factory settings or when further adjustments are required.
Detection of the inflection point
The logic of detecting the inflection point is given below.
Assuming that the remainder obtained by subtracting the amount of change C of the
indicated value over interval A from the amount of change of the indicated value over
interval B is D, when the amount of change D exceeds the inflection point judgment
value, point a is held as an inflection point.
If there are two or more inflection points in the hold period, the point having a larger
change is held.
The inflection point is normally detected by A=B, but it may easily be detected with
A<B where the slope is gradual.
52

7-3-1. Minimum Slope

How to set
How to set
Setting call Page 2
Hold
Setting call Page 2
Hold
The interval setting is the number of samplings. Since the F371's
sampling speed is 2000 times/sec., one sampling is 0.5msec.
Therefore, setting of the interval at 100 means setting of 50msec.
Load change time
True inflection point
Inflection point
Setting range: 00001 - 99999

7-3-2. Interval A and Interval B

Setting range:
- Interval A+Interval B 1000
- 10 Interval A(B) 990
7.HOLD FUNCTIONS
Caution regarding inflection point hold
If the detection interval A and B are set
too short, fine load changes may be
detected as shown in the illustration on
the left-hand side, so that a correct value
cannot be held.
In this case, set the detection interval B
large enough to bring it as close to the
load change time as possible, and also
set the inflection point judgment value
large according to the amount of change
at that time, so that the inflection point
is held at a correct position.
53
7.HOLD FUNCTIONS
100
20
10
0
100msecA100msec
B
D = 80
C = 10
Example of inflection point hold setting
Example of ideal waveform setting
Set the load change time (between the inflection point and when the change stops) to
interval B.
In the example, it is set to 200 since it is 100 msec.
Set the same value as interval B to interval A.
Set load D obtained by subtracting load C changing with interval A from the load
changing with interval B to the inflection point judgment value.
In the example, set load change D=80 obtained by subtracting load C=10 changing
with A from load 90 changing with B to the inflection point judgment value.
However, since an inflection point is not judged until load change D exceeds the
inflection point judgment value, actually set the inflection point judgment value a
slightly smaller than D.
If the inflection point cannot be located successfully
When holding above the inflection point and
moving downward
 1) It is considered that the inflection point
judgment value is small with respect to load
change D. Set the inflection point judgment
value larger.
 2) If lowering is insufficient in 1), increase
interval A.
54
When holding below the inflection point and
moving upward Interval B is too long and the
inflection point judgment value is too large.
Shorten interval B and decrease the inflection
point judgment value.

7-4. Hold Setting --- mean value ---

How to set
Setting call Page 2
Hold
About the maximum mean value detection time
Although the detection period is specified by the H/M signal, etc.,
detection cannot be carried out exceeding the maximum mean value
detection time set according to the mean value sampling count. If the
maximum mean value detection time is exceeded, detection ends
automatically, when the mean value is held.

7-4-1. Mean Value Sampling Count

If the mean value sampling count is set at 2 or more in the mean value hold, the
representative value of the sampling values by the set count (mean value by the count) is
adopted as the sampling data used for mean value calculation.
The maximum mean value detection time with the setting “1” is 5 sec., but the mean
value detection time can be extended by this setting.
Maximum mean value detection time
= Mean value sampling count × 5 [sec.],
where the number of updates of the mean value will decrease.
7.HOLD FUNCTIONS
Number of updates of the mean value
= 2000 / Mean value sampling count [times/sec.]
Setting range: 1 - 999Times
55
7.HOLD FUNCTIONS
Indicated Value
t2
t3
t1
Sensor Input Value
Detection・Hold Period
Decide
t
T/H
HI-LO Limit Judging Output
H/E
OFF
ON
OFF
ON
OFF
ON
HOLD
OFF
ON
Button
HOLD Status Display
Status Display
ONBlink Blink
OFF
PEAK

7-5. Hold Operation

7-5-1. Sample Hold

When the T/H signal is inputted, a desired point is held, and the H/E output is turned on.
Hold of the value continues as long as the T/H signal is on.
t1: A delay time between the instant when the T/H signal is inputted and the instant
t2A delay time between the instant when the T/H signal is released and the instant
t3:A minimum reset signal width required for releasing the hold
when the indicated value is held
1.0ms (max.)
when the indicated value returns to tracking
1.0ms (max.)
1.0ms (min.)
56

7-5-2. Peak Hold

Indicated Value
t2
t3
Sensor Input Value
Detection
t
T/H
HI-LO Limit Judging Output
H/E
Detection
t1t2t1
t3
Note: During the undetermined period, the judging output varies with
fluctuations in the input waveform. However, the H/E output
remains on during the undetermined period. Read the judging
result when the indicated value becomes stable (immediately
before the T/H signal rises).
OFF
ON
OFF
ON
OFF
ON
OFF
ON
Status Display
PEAK Status Display
ON
Blink Blink
OFF ON
OFF
OFF
OFF
OFF
OFF
HOLD Button
Undetermind Period(Note)
Undetermind Period(Note)
Hold Period
Hold Period
HOLD
The maximum value (peak) in the positive direction of the specified period is held.
The period is specified by the setting of “all period”, “external signal”, “time”, or “time
with trigger”.
7.HOLD FUNCTIONS
(Example) All Period Peak Hold
t1: A delay time between the instant when the T/H signal is inputted and the instant
when the indicated value is held
1.0ms (max.)
t2: A delay time between the instant when the T/H signal is released and the instant
t3:A minimum reset signal width required for releasing the hold
when the indicated value returns to tracking
1.0ms (max.)
1.0ms (min.)
57
7.HOLD FUNCTIONS
Indicated Value
t2
t3
Sensor Input Value
Detection
t
T/H
HI-LO Limit Judging Output
H/E
Detection・Hold Period
t1t2t1
t3
Note: During the undetermined period, the judging output varies with
fluctuations in the input waveform. However, the H/E output
remains on during the undetermined period. Read the judging
result when the indicated value becomes stable (immediately
before the T/H signal rises).
OFF
ON
OFF
ON
OFF
ON
OFF
ON
HOLD Status Display
PEAK Status Display
ON
Blink Blink
OFF ON
OFF
OFF
OFF
OFF
OFF
HOLD Button
Undetermind Period(Note)
Undetermind Period(Note)
Hold Period

7-5-3. Valley Hold

The maximum value (valley) in the negative direction of the specified period is held.
The period is specified by the setting of “all period”, “external signal”, “time”, or “time
with trigger”.
(Example) All Period Valley Hold
t1: A delay time between the instant when the T/H signal is inputted and the instant
when the indicated value is held
1.0ms (max.)
58
t2: A delay time between the instant when the T/H signal is released and the instant
when the indicated value returns to tracking
1.0ms (max.)
t3:A minimum reset signal width required for releasing the hold
1.0ms (min.)

7-5-4. Peak-to-Peak (P-P) Hold

Indicated Value
Sensor Input Value
Detection・Hold Period
Undetermind
t
T/H
HI-LO Limit Judging Output
H/E
t2
t1
t3
Standard Line
(Hold Section Only)
Indicated Value = 0
Indicated Value=0
OFF
ON
OFF
ON
OFF
ON
HOLD
OFF
ON
Button
HOLD Status Display
PEAK Status Display
ON
Blink
OFF
OFF
OFF
OFF
Period(Note)
Note: During the undetermined period, the judging output varies with fluctuations in the
input waveform. However, the H/E output remains on during the undetermined
period. Read the judging result when the indicated value becomes stable
(immediately before the T/H signal rises).
The maximum value of the difference between the peak and valley over the specified
period is held.
The period is specified by the setting of “all period”, “external signal”, “time”, or “time
with trigger”.
(Example) All Period Peak-to-Peak (P-P) Hold
7.HOLD FUNCTIONS
t1: A delay time between the instant when the T/H signal is inputted and the instant
when the indicated value is held
1.0ms (max.)
t2: A delay time between the instant when the T/H signal is released and the instant
when the indicated value returns to tracking
1.0ms (max.)
t3:A minimum reset signal width required for releasing the hold
1.0ms (min.)
59
7.HOLD FUNCTIONS
Indicated Value
t3
Sensor Input Value
Detection
t
HI-LO Limit Judging Output
Hold Period
Decide
H/M
t2
t1
Auto Start Level
T/H
H/E
OFF
ON
OFF
ON
OFF
ON
OFF
ON
HOLD
PEAK
ON
Blink ON
OFF
ON
OFF
OFF
OFF
HOLD
Buttun
OFF
ON
or
OFF
Comes on if a value is held with conditions met in the
detection period, and goes out if a value is not held.

7-5-5. Relative Maximum and Relative Minimum Hold

Detection starts when the indicated value crosses the auto start level with the H/M signal
inputted. Detection is performed as long as the H/M signal is on.
The hold is released by turning on the T/H signal as a reset signal.
The H/E output signal is on between the instant when the hold starts and the instant when
the T/H signal is turned on.
If the auto start level is not needed as a starting condition of the detection period, set
99999.
(Example) Relative Maximum Hold
t1: A delay time between the instant when the indicated value exceeds the auto start
level and the instant when the value to be held is detected
0.5ms (max.)
t2A delay time between the instant when the H/M signal is inputted and the instant
when the value to be held is detected
1.0ms (max.)
t3:A minimum reset signal width required for releasing the hold
1.0ms (min.)
60
7.HOLD FUNCTIONS
The Hold Function in General
The delay times of the judging signal and the H/E signal do not include a delay
time of the analog circuit (low-pass filter). Also, the delay times are calculated
without the digital filter. Since the judging signal and the H/E signal are
generated for the value produced through the analog and digital filters,
transmission of the signals becomes slower and the delay of each output
signal increases with enhancement of the filters.
If the H/M signal is continuously input, the previously held value is reset by
the on timing, and hold operation newly starts.
61
7.HOLD FUNCTIONS
Indicated Value
t3
Sensor Input Value
Detection
t
HI-LO Limit Judging Output
Hold Period
Decide
H/M
t2
t1
Auto Start Level
T/H
H/E
OFF
ON
OFF
ON
OFF
ON
OFF
ON
HOLD
PEAK
ON
Blink ON
OFF
ON
OFF
OFF
OFF
HOLD
Buttun
OFF
ON
or
OFF
Comes on if a value is held with conditions met in the
detection period, and goes out if a value is not held.

7-5-6. Inflection Point Hold

Detection starts when the indicated value crosses the auto start level with the H/M signal
inputted. Detection is performed as long as the H/M signal is on.
The hold is released by turning on the T/H signal as a reset signal.
The H/E output signal is on between the instant when the hold starts and the instant when
the T/H signal is turned on.
If the auto start level is not needed as a starting condition of the detection period, set
99999.
(Example) Inflection Point Hold
t1: A delay time between the instant when the indicated value exceeds the auto start
level and the instant when the value to be held is detected
0.5ms (max.)
t2A delay time between the instant when the H/M signal is inputted and the instant
when the value to be held is detected
1.0ms (max.)
t3:A minimum reset signal width required for releasing the hold
1.0ms (min.)
62

7-5-7. Mean Value Hold

Real-time
Hold values
t
T/H
Judging Outpu
H/E
Decide
H/M
OFF
ON
OFF
ON
OFF
ON
OFF
ON
Indecided
t1 t1 t1
Detection
Hold Period
values
The detection period is limited by the maximum mean value detection
time set according to the “mean value sampling count”.
For details, see the section on Mean Value Sampling Count page 55.
The mean value of the sampling values over the specified period is calculated and
updated to perform comparison operation.
The period is specified by the setting of “all period”, “external signal”, “time”, or “time
with trigger”.
(Example) Externally Specified Period Mean Value Hold
7.HOLD FUNCTIONS
t1A delay time between the instant when the T/H or H/M signal is inputted and the
instant when operation is performed.
1.0mSMAX.
63
7.HOLD FUNCTIONS
Indicated Value
t2
t3
Sensor Input Value
Detection
t
T/H
HI-LO Limit Judging Output
H/E
Detection
t1t2t1
t3
Note: During the undetermined period, the judging output varies with
fluctuations in the input waveform. However, the H/E output
remains on during the undetermined period. Read the judging
result when the indicated value becomes stable (immediately
before the T/H signal rises).
OFF
ON
OFF
ON
OFF
ON
OFF
ON
Status Display
PEAK Status Display
ON
Blink Blink
OFF ON
OFF
OFF
OFF
OFF
OFF
HOLD Button
Undetermind Period(Note)
Undetermind Period(Note)
Hold Period
Hold Period
HOLD

7-6. How to Specify the Hold Detection Period

7-6-1. All Period

By this method, the hold detection period is externally specified by the T/H signal.
Detection starts with the T/H signal ON to perform each hold operation.
Detection ends with the T/H signal OFF to reset the hold.
(Example) All Period Peak Hold
t1: A delay time between the instant when the T/H signal is inputted and the instant
when the indicated value is held
1.0ms (max.)
t2: A delay time between the instant when the T/H signal is released and the instant
when the indicated value returns to tracking
1.0ms (max.)
t3:A minimum reset signal width required for releasing the hold
1.0ms (min.)
64
7-6-2. Externally Specified Period Hold
Indicated Value
t3
Sensor Input Value
Detection
t
T/H
HI-LO Limit Judging Output
H/E
Hold Period
t2t1
Decide
H/M
OFF
ON
OFF
ON
OFF
ON
OFF
ON
OFF
ON
HOLD
PEAK
ON
Blink ON
OFF ON
OFF
OFF
OFF
HOLD Button
(Peak, Valley , Peak-to-Peak and Mean Value)
By this method, the hold detection period is externally specified by the H/M signal to
maintain the hold value until the reset signal is turned on.
The hold is released by turning on the T/H signal as a reset signal.
The H/E output signal is on between the instant when the H/M signal is turned off and
the instant when the T/H signal is turned on.
(Example) Period Specified Peak Hold
7.HOLD FUNCTIONS
t1A delay time between the instant when the H/M signal is inputted and the instant
when the value to be held is detected
1.0ms (max.)
t2A delay time between the instant when the H/M signal is released and the instant
when the value to be held is determined
1.08ms (max.)
t3:A minimum reset signal width required for releasing the hold
1.0ms (min.)
65
7.HOLD FUNCTIONS
Indicated Value
t3
Sensor Input Value
Detection
t
HOLD
HI-LO Limit Judging Output
H/E
Hold Period
t2t1
Decide
H/M
T/H
Time
OFF
ON
OFF
ON
OFF
ON
OFF
ON
OFF
ON
OFF
ON
HOLD
PEAK
ON
Blink ON
OFF ON
OFF
OFF
OFF
HOLD Button
7-6-3. Time Specified Period Hold
(Peak, Valley, Peak-to-Peak and Mean Value)
Hold is detected during the predetermined time (hold detection time) from the point in
time when the H/M signal is turned on. The hold is released by turning on the T/H signal
as a reset signal.
The H/E output signal is on between the instant when the H/M signal is turned off and
the instant when the T/H signal is turned on.
(Example) Time Specified Period Peak Hold
t1A delay time between the instant when the H/M signal is inputted and the instant
when the value to be held is detected
1.0ms (max.)
t2A delay time between the instant when the hold detection time is expired and the
instant when the value to be held is determined
1.0ms (max.)
t3:A minimum reset signal width required for releasing the hold
1.0ms (min.)
66
7-6-4. Time Specified Period Hold with Trigger
Indicated Value
t3
Sensor Input Value
Detection
t
HOLD
HI-LO Limit Judging Output
H/E
Hold Period
t2t1
Decide
T/H
Time
Auto Start Level
OFF
ON
OFF
ON
OFF
ON
OFF
ON
OFF
ON
HOLD
PEAK
ON
Blink ON
OFF ON
OFF
OFF
OFF
HOLD Button
(Peak, Valley, Peak-to-Peak and Mean Value)
Hold is detected during the predetermined time (hold detection time) from the point in
time when the indicated value crosses the auto start level. The hold is released by turning
on the T/H signal as a reset signal.
The H/E output signal is on between the instant when the hold detection time ends and
the instant when the T/H signal is turned on.
(Example) Time Specified Period Peak Hold with Trigger
7.HOLD FUNCTIONS
t1: A delay time between the instant when the indicated value exceeds the auto start
t2: A delay time between the instant when the hold detection time is expired and the
t3:A minimum reset signal width required for releasing the hold
level and the instant when the value to be held is detected
0.5ms (max.)
instant when the value to be held is determined
1.0ms (max.)
1.0ms (min.)
67

8.MULTI-HOLD FUNCTION

I/O terminal block
B1 COM2
B2 CODE0
B3 CODE1
B4 CODE2
B5 CODE3
Caution
It takes 15msec at the maximum for the
changed CH to become effective. During
this period, which CH is used for
comparison is undefined.
Also, when the CH is switched, the hold and
graph functions are reset to wait for starting
with the after-switching CH conditions
irrespective of the previous operation.
8. MULTI-HOLD FUNCTION
By this function, up to 16 types of hold, graph and comparison set values can be stored
and selected with external switching signals of CODE0 - CODE3.
Normally, if there is no entry for CODE0 - CODE3, the set value of CH0 is selected, but
when CODE0 - CODE3 are in the following conditions, the set value of each setting CH
is selected.
CODE3 CODE2 CODE1 CODE0 SettingCH
0 0 0 0 SettingCH00
0 0 0 1 SettingCH01
0 0 1 0 SettingCH02
0 0 1 1 SettingCH03
0 1 0 0 SettingCH04
0 1 0 1 SettingCH05
0 1 1 0 SettingCH06
0 1 1 1 SettingCH07
1 0 0 0 SettingCH08
1 0 0 1 SettingCH09
1 0 1 0 SettingCH10
1 0 1 1 SettingCH11
1 1 0 0 SettingCH12
1 1 0 1 SettingCH13
1 1 1 0 SettingCH14
1 1 1 1 SettingCH15
 (0: open, 1: short)
68

8-1. About Changing of the Setting CH

How to set
For making the settings of all CHs equal, set the CH No. to CH16.
The value set with CH16 is set for all CH00 CH15.
NOTE
No control CH can be designated with the CH change key.
Designate control CH(s) with external selector signals
CODE0 - CODE3.
When changing the hold or comparison set value of each CH, set the CH No. with the
CH change key on the mode selection screen, and change the set value.
1Select the setting CH on the mode setting screen.
8.MULTI-HOLD FUNCTION
2Then, enter each set value in a likewise manner.
69

9.WAVEFORM DISPLAY

X-axis (time axis)
Graphic display START/STOP
Cursor move keys (The cursor moves on the X-axis.)
Cursor ON/OFF key
Y-axis (load axis)
Caution
When the cursor display is on and when the setting screen is
open, the graph is not upgraded.
Indicated value display (At cursor ON time, the cursor
          point (green) value is displayed.)
9. WAVEFORM DISPLAY

9-1. Graphic Display Screen

Graph is updated while usually operating on the display screen (a comparison ,a hold
,graph ).
70

9-1-1. Hold Point Plotting

Held point
Example) Relative maximum hold
Plotting Range
D Z
100
Y
X
200
X End Point
Y End Point
Y Start Point
0
By using the hold function and the graph plotting function together, the held point (red)
can be plotted.

9-1-2. X-axis and Y-axis on the Graph Plotting Screen

9.WAVEFORM DISPLAY
X-axis
Y-axis
The X-axis is a time setting axis. One graph screen is plotted between the start signal
input point and the X start point.
There are 200 plotting points, and typical values for the predetermined time divided by
this number of plotting points are plotted.
The Y-axis is a load setting value. A graph is plotted between the Y start point and the Y
end point. (There are 100 plotting points.)
71
9.WAVEFORM DISPLAY
t
START/STOP
ST/SP
Star t
Plotting Period
Screen Hold
Plotting Period
X end point
X end point
Screen Clear
Interval Time
Key
Outside Input
time setteing
time setteing

9-2. Graph Plotting

9-2-1. Continuous

Graph plotting is started by turning on the START/STOP key or turning on the ST/SP
external input. When it ends on one screen, the screen is cleared after expiration of the
interval time and graph plotting restarts on the next screen. It is ended by turning off the
START/STOP key or turning off the ST/SP external input.
72

9-2-2. External

t
START/STOP
ST/SP
Key
Plotting
X end point time setteing Screen Hold
Period
Trigger wait
1.0msec more
Plotting is started by turning on the START/STOP key or turning on the ST/SP external
input.
Plotting ends on one screen at the predetermined time of the X end point.
9.WAVEFORM DISPLAY
73
9.WAVEFORM DISPLAY
t
X end point time setteing
Screen Hold、Trigger wait
Trigger Level
Plotting
Period
t
ST/SP
Outside Input
X end point time setteing
Trigger Level
Level Detection Period
Plotting
Period
Screen Hold、Trigger wait
1.0msec more

9-2-3. Level

Plotting is started when the conditions of the detection mode are met in comparison of
the trigger level set value and indicated value.
Plotting ends on one screen at the predetermined time of the X end point.

9-2-4. External + Level

Plotting is started according to the conditions of the level detection mode in comparison
of the trigger level and indicated value after the ST/SP external input is turned on.
Plotting ends on one screen at the predetermined time of the X end point.
(Example) Level Detection Mode Passing under.
74

9-3. Graph Function

How to set
How to set
How to set
Setting call Page 1
Graph
Setting call Page 1
Graph
Setting call Page 1
Graph
Set the graph plotting mode.
Modes
Continuous, external, level, external + level

9-4. Interval Time

9.WAVEFORM DISPLAY
If you select "continuous" in the graph function setting, set the graph plotting operation
interrupting time from clearing the screen until moving to the next graph plotting
operation. During this time, the graph screen is held.
Setting range: 00.0 - 99.9 sec.

9-5. Graph Start Level

If you select [level] or [external + level] in the graph function setting, set the graph
plotting start level.
Setting range: -99999 - 99999
75
9.WAVEFORM DISPLAY
How to set
Setting call Page 1
Graph
- Passing
Graph plotting starts when the indicated value crosses the level set
value.
- Passing over
Graph plotting starts when the indicated value crosses the level set
value upward.
- Passing under
Graph plotting starts when the indicated value crosses the level set
value downward.
- Large
Graph plotting starts when the indicated value is larger than the level
set value.
- Small
Graph plotting starts when the indicated value is smaller than the
level set value.

9-6. Level Detection Mode

If you select [level] or [external + level] in the graph function setting, set the graph
plotting start level comparison conditions.
Conditions: passing, passing over, passing under, large, small
76

9-7. X Start Point

How to set
Setting call Page 1
Graph
Press the X button on the graph display to go direct to the X end point entry screen.
Simple setting call
Set the time to display by one screen in the range between 0.1 sec. and 99.9 sec.
9.WAVEFORM DISPLAY
77
9.WAVEFORM DISPLAY
How to set
Setting call Page 2
Graph
Press the Y button on the graph display to go direct to the Y end point entry screen.
Simple setting call

9-8. Y (Load) Start Point and Y (Load) End Point

Setting range: -99999 - 99999 (where Y start point < Y end point)
78

10.VOLTAGE OUTPUTVOL OUT

+
-
F. G
External
- +
F371 inside
equipment
10. VOLTAGE OUTPUTVOL OUT
This interface extracts analog voltage proportional to sensor input signals. This interface
is convenient for observation and recording of waveforms with a recorder, etc.,
connected.
The output level is approx. 2V per 1mV/V of sensor input.
Example of output equivalent circuit and external equipment connection
Output signals are not the indicated values themselves because they are extracted before
sensor input signals are A/D-converted.
Therefore, output signals do not coincide with the digitally processed indicated values,
such as the digital zero and digital filter.
79

11.SCREEN LOCK

Screen lock status
Event output
OFF
ON
200msec
The end of graph plotting
11. SCREEN LOCK
When terminals are short-circuited, key operation is invalidated on the measurement
screen (comparison/hold/graph). However, the measurement screen can be changed
(hold to graph) and the cursor function is valid. During lock, the lock status is displayed
on the left-hand side of the screen, and the alarm sounds if any invalid key is pressed.

12. EVENT OUTPUT AT THE END OF GRAPH PLOTTING

Each time a graph is plotted on screen, a pulse signal is outputted (the pulse width is
200msec). The waveform can automatically be taken in through the RS-232C interface
by connecting to the event input of the DS400.
80

13. RS-232C INTERFACE

The RS-232C is an interface to read the indicated value and status of the F371 and to
write parameters into the F371.
This interface is convenient to process controls, totals, records, etc., by connecting the
F371 to a computer, process controller, sequencer or the like.
13.RS-232C INTERFACE

13-1. Communication Specifications

13-1-1. Specifications

Signal level Based on RS-232C
Transmitting distance:Approx.15m
Transmitting method :Asynchronous,Full duplex
Transmitting speed :1200,2400,4800,9600 or 19200bps Selectable
Bit configuration Start bit 1
Character length 7 or 8 bit Selectable
Stop bit 1 or 2 bit Selectable
Parity Parity noneodd or even Selectable
Terminator CRCR+LF Selectable
Code ASCII
81
13.RS-232C INTERFACE
Case
F.
TxD
RxD
DTR
GND
GND
DTR
TxD
RxD
CD
DSR
CTS
RTS
F. G
TCP8080-01-520
F371
D-sub 9 pin
Personal computer
, etc.
Cabling diagram
CA371-232 (optional)

13-1-2. Connector Pin Assignment

This connector connects the RS-232C.
Pin No. Signal name
1
2TxD
3RxD
4DTR
5
6
7GND
8
Case F.G.

13-1-3. Cable

The following shows an example of connection between DTE-DTE terminals. This will
require modification depending on the equipment to be connected. For details, see the
instruction manual of the equipment to be connected.
82
* This connection diagram shows an example of connecting a DTE (data terminal
equipment) personal computer.
Use a straight cable for connecting a DCE (data circuit terminating equipment) such
as a modem.
* Before preparing a cable, check the connector shape and pin assignment of the
equipment to be connected again.

13-2. RS-232C Interface Setting

How to set
How to set
How to set
Setting call Page 1
Option
Setting call Page 1
Option
Setting call Page 1
Option
Set the RS-232C communication conditions of the F371.

13-2-1. Communication Mode

Communication mode 0, communication mode 1, communication mode 2

13-2-2. Baud Rate

13.RS-232C INTERFACE
120024004800960019200bps

13-2-3. Character Length

7bit8bit
83
13.RS-232C INTERFACE
How to set
How to set
How to set
Setting call Page 1
Option
Setting call Page 1
Option
Setting call Page 2
Option

13-2-4. Parity Bit

None, odd, even

13-2-5. Stop Bit

1 bit, 2 bit

13-2-6. Terminator

CRCR LF
84

13-3. Communication Modes

Host
F371
RA CR
RA
Ter mi nat or
1 0 0 0 0 .
Host
F371
RD CR
R D 0 0 0 0 0 0 0
1:ON 0:OFF
1:ON 0:OFF
Near Zero 1:ON
Output signal 0:OFF
HI 1:ON
Output signal 2:OFF
L O 1:ON Output signal 0:OFF
OK 1:ON
Output signal 0:OFF
Undefined
Hold
Stable
Ter min at or
1.Communication Mode 0
Communications are carried out by commands from the host computer.
The indicated value, status and parameters can be read, and parameters can be written.
2. Communication Mode 1
The indicated value and status are continuously transmitted.
3. Communication Mode 2
The indicated value is transmitted upon printing.
13.RS-232C INTERFACE

13-4. Communication Format

1. Communication Mode 0
- Reading the indicated value (sign, 5-digit indicated value, decimal point)
- Reading the status (7-digit)
The hold bit is operated at the same timing as the H/E signal.
85
13.RS-232C INTERFACE
Host
F371
RC CR
R C 0 0 0 0 0 0 0
1:ON 0:OFF
1:ON 0:OFF
1:ON
0:OFF
1:ON 2:OFF
LL 1:ON Output signal 0:OFF
HH 1:ON Output signal 0:OFF
Undefined
LOAD
Ter min at or
LOAD
OVER
OVER
1
W
LF2
CR LF3 1
W
CR LF41
W
CR
CR LF05 1
W
CR LF8 4
W
CR LF6 1
W
0
0
1
W
LF1CR±
±
±
±
±
CR LF44W00000
Set Value Lock)
Calibrationt Value
Lock)
Calibrationt Value
Lock)
Set Value Lock)
Set Value Lock)
Set Value Lock)
Set Value Lock)
Set Value Lock)
- Writing parameters
HI-HI Limit
HI limit
LO limit
LO-LO Limit
Hysteresis
Calibration Value Selection
Digital Offset Setting
Near Zero
86
Hold mode
CR LF1 2
W
0 0 0 0
CR LF2 2
W
0 0
CR LF32
W
±
CR LF42
W
000
CR LF52
W
00 0 00
CR LF62
W
0
CR LF72
W
0
00
CR LF82
W
0
00
3
W
LF1 CR
0 0000
3
W
LF2 CR
0 00
CR LF18
W
0
0 0 0
0
0
CR LF 043
W
0000
CR LF3 3 W ±
CR LF 0F1
W
000
Set Value Lock)
1 999
0 3
010 120
010 120
Set Value Lock)
Set Value Lock)
Set Value Lock)
Set Value Lock)
2
W
LF9 CR
0 00
Hold Section Setup
Hold Detect Time
13.RS-232C INTERFACE
Auto Start Level
Minimum Count
valley Detect Level
Minimum Slope
IntervalA
IntervalB
Graph Function
Interval Time
Graph Start Level
Level Detect Mode
CH
Hold Point Shift
87
13.RS-232C INTERFACE
Hold mode
0tracking 5relative maximum
1sample 6relative minimum
2peak 7inflection point
3valley 8mean value
4peak-to-peak
Hold period setting
0all period
1external signal
2time
3time with trigger
Valley detection level
01/4 times 41.25 times
11/2 times 51.5 times
23/4 times 62 times
31 time 73 times
Graph mode
0continuous
1external
2level
3external + level
Level detection mode
0passing
1passing over
2passing under
3large
4small
Calibration Value Selection
0calibration value 0
1calibration value 1
2
calibration value 2
3calibration value 3
Set value correspondence table
88
13.RS-232C INTERFACE
Host
F371
G
R
CR
Decimal Point and Five digits Data1
00 19
R
Data Section No.
G
±
00 19
Data Section No.
……
Teaminator
±±
……
Decimal Point and Five digits Data9 Decimal Point and Five digits Data10
HOST
H
R
R
H
……
CR
ターミネータ
10 0 00025
DATA000 ~ 199
F371
One graph plotting screen includes 200 pieces of data. In reading
waveform data, data can be read by being divided into 20 sections
(by 10 pieces).
Data Section No.
Y
DZ
X
Plotting Range
Data 0
Data 199
Section 00 Data 0 Data 9 Section 01 Data 10 Data 19
Section 19 Data 190 Data 199
- Reading waveform data
Wave form hold point data read-out(data No, a mark, a decimal point,5 figures of
directions value)
89
Loading...