● The F350 is a digital indicator that is capable of two-channel inputs of strain gauge
sensors.
● Calculations, such as addition and subtraction, and comparison judgments are made
by using the input values of two channels. The F350 is also suitable for balance
measurement.
● The F350 is equipped with comparison judgments on the input value of each channel,
and analog voltage output.
● The A/D conversion speed of each channel and the digital processing speed for
calculations and judgments are as high as 3,000 times/sec. It can also be switched to
300 times/sec.
1 OUTLINE
Chapter
1
OUTLINE
● The F350 is equipped with hold functions with respect to calculated values.
(sample, peak, bottom, and average holds)
● The F350's compact size of DIN96×96 realizes space-saving.
● By mounting an optional interface according to the purpose, ease of load control can
be further improved.
1
1
CAUTION
F350 body・・・1F350 operation manual・・・1
AC input cord・・・1
Conversion plug for
Mini screwdriver for Connector for BCD parallel
(with BCD parallel data output option)
AC input code・・・1
(with D/A converter option)
Ferrite core・・・2
(with unity band)
External input/output
(with solderless terminal)
connector・・・1
terminal block connection・・・1
data output・・・1
1 OUTLINE
Chapter
1
OUTLINE
1-2.Contents of the package
The packaging box contains the following.
Be sure to check them before use.
CAUTION
The attached AC cable is intended for domestic use in Japan, and it is rated as 125V AC, 10A. If it
is used at voltages exceeding the rating or it is used overseas, have a suitable AC cable ready
separately.
2
2
CAUTION
Strain gauge type sensor
(Option)
D/A Converter
Valve, recorder
(Option)
(Option)
PrinterDisplayVarious converters
PC
* +.1* +.1
* +. 11 -
* 1 . &
2 ' # -
%J
%J
(0%'5%
*+.1
<'41
*1.&
F
350
Main status display
Sub status display
Setting key pad
Main numerical display
Sub numerical display
section
section
section
section
1 OUTLINE
1-3.About connectable devices
Chapter
1
OUTLINE
1-4.Appearance description
3
■Front panel
3
CAUTION
In this operation manual, the display status is expressed as follows:
Lighting:Status display section → Black
Numerical display section → White
Blinking:Lighting state + Lines in a radial pattern
Being unlit:Status display section → Gray
Numerical display section → Nothing displayed
LO
H
I
Ch
1
Ch 2
OK
LO
PEAK
HOLD
H
I
LO
H
I
Lighting
Blinking
Being unlit
Chapter
1
OUTLINE
1 OUTLINE
Main status display section
The status with respect to calculated values is indicated.
HI:Lights when the indicated value is larger than the calculation HI limit.
Indicates that the external output "calculation HI" is ON.
OK:Lights when the indicated value is smaller than the calculation HI limit and larger
than the calculation LO limit.
Indicates that the external output "calculation OK" is ON.
LO:Lights when the indicated value is smaller than the calculation LO limit.
Indicates that the external output "calculation LO" is ON.
PEAK:Blinks during hold detection, and lights when a hold is confirmed.
However, it is always unlit in all section sample hold.
HOLD:Lights during hold detection and when a hold is confirmed.
However, it blinks when hold waiting in all section sample hold.
Main numerical display section
The following display patterns exist.
Indicated value:Displays calculation results during measurements.
Error display:Displays an error message when an error occurs.
Setting value: Displays setting values in setting mode.
Sub status display section
The status with respect to ch1/ch2 values is indicated.
HI:Lights when the indicated value is larger than the ch1/ch2 HI limit.
Indicates that the external output "ch1/ch2 HI output" is ON.
LO:Lights when the indicated value is smaller than the ch1/ch2 LO limit.
Indicates that the external output "ch1/ch2 LO output" is ON.
Sub numerical display section
The following display patterns exist.
Indicated value: Displays ch1/ch2 values during measurements.
Error display: Displays an error message when an error occurs.
Setting value:Displays the mode number and item number at the ch1 display section in
setting mode; the currently-selected weight value may be displayed at the ch2
display section according to the setting item
4
4
CAUTION
FNC
ESC
HI/LO
ZERO
HOLD
Setting key pad
These are keys for commanding settings and operations.
・Use this key to go to a setting mode state.
・Use this key to cancel setting/execution.
・Use this key to go back from a setting mode state to an indication display state.
・Use this key to take a shortcut from an indication display state to HI/LO limit setting.
・
Use this key to go from an indication display state to ch1 or ch2 Digital Zero ready.
・Use this key to execute/release the calculated value hold in an indication display state.
・Use this key to switch between ch1 setting and ch2 setting, etc., in a Digital Zero
ready state or setting item select state.
・Use this key to set whether or not to sign a setting value, when available, in a setting
value input state.
1 OUTLINE
Chapter
1
OUTLINE
・Use this key to select a trailing setting item number in a setting mode state.
・Use this key to select the previous item number in a setting item select state.
・Use this key to increment the numerical value by one in a setting value input state.
・Use this key to select a leading setting item number in a setting mode state.
・Use this key to select the next item number in a setting item select state.
・Use this key to decrement the numerical value by one in a setting value input state.
・Use this key to execute Digital Zero reset in a Digital Zero ready state.
・Use this key to increment the leading setting mode number by one in a setting mode
state.
・Use this key to move the setting digit by one in the lower direction in a setting value
input state.
・Use this key to execute Digital Zero in a Digital Zero ready state.
・Use this key to move to a setting value input state in a setting item select state.
・Use this key to validate the setting value in a setting value input state.
5
5
CAUTION
Options slot
Power input terminal
Protective ground
Signal input/output
Analog input/output
connector
connector
Frame ground
Chapter
1
OUTLINE
1 OUTLINE
■Rear panel
Analog input/output connector
Connect a strain gauge sensor or voltage output.
* For connection, see "
■ Analog input/output connector connection" on page 10.
Signal input/output connector
Connect external input/output signals or SI/F (serial interface).
* For connection, see "
■ External I/O connection" on page 14.
Power input terminal block
Connect an AC input cord.
The input power source is free in the range of 100 to 240V AC (-15%, +10%). The frequency is 50/
60Hz.
* For connection, see "
■ Power input terminal connection" on page 9.
Protective ground
Be sure to ground the protective ground terminal to prevent electric shocks.
Frame ground
6
6
Please ground the frame ground terminal to prevent failures due to static electricity.
(The frame and the frame ground terminal are conducted.)
It may be better to remove depending on the environment of the installation location.
Options slot
One option board can in stall in the option slot.
・BCD parallel data output(BCO)
・RS-232C interface(232)
・D/A converter voltage output(DAV)
・D/A converter current output (DAI)
2INSTALLATION & CONNECTION
2I
NSTALLATION & CONNECTION
The following are precautions related to connection to the signal input/output terminal block.
The precautions described here are important for safety.
Make connections after properly understanding the description.
WARNING
● Do not connect commercial power directly to the signal input/output terminals.
● Make connection to the signal input/output terminals with no power applied
because you may receive an electric shock.
● For connections to the signal input/output terminals, carry out wiring properly after
confirming the signal names and pin assignment numbers.
● Use of the F350 is limited to category II specified by EN61010. Overvoltage applied
to the signal input/output terminals should not exceed the value defined to category
II.
● After wiring, make sure to mount the attached terminal block cover. Otherwise, you
may receive an electric shock.
● Turn on the power after carefully checking the wiring, etc.
Chapter
2
INSTALLATION & CONNECTION
● Do not touch the signal input/output terminals while the power is on. Otherwise, you
may receive an electric shock, or a malfunction may occur.
● The F350 conforms with EMC Directives as a product used in industrial
environments (class A). If it is used in living environments, radio disturbance may
occur. In that case, take suitable measures.
CAUTION
● Tighten the terminal screws at specified torque. If the terminal screws are loose,
short-circuiting, fire, or a malfunction may occur.
● Use shielded cables.
7
7
2 INSTALLATION & CONNECTION
92mm
Panel-cut
+1
0
–
Panel thickness
1.6 to 3.2mm
dimensions
92mm
+1
0
–
2-1.Installation
To install the F350 into a control panel, use the following procedure.
1. Make a hole in the panel according to
Chapter
2
INSTALLATION & CONNECTION
the panel-cut dimensions.
2. Remove the screws (two), and remove
the guide rails from both sides.
* Do not use other screws than those installed to
the F350 body.
3. Fit in the F350 from the front of the panel.
4. Install from the rear the guide rails having
been removed from both sides in Step 2,
and fix them with the screws (two).
8
8
2-2.Connection
Within 6mm
Black
White
Green/Yellow
Color description are for the attached
AC input cord.
■Power input terminal connection
Connect an AC input cord. The input power source is free in the range of 100 to 240V AC (-15%,
+10%).
The frequency is 50/60Hz.
Make connections to the terminal block with a solderless terminal within 6mm in diameter as
shown in the illustration so as not to let the tip of the cable spread out.
1. Make sure that no power is applied.
2. Remove the screws (two), and remove
the terminal block cover.
3. Remove the two screws(M3) at
the terminal block
.
2 INSTALLATION & CONNECTION
Chapter
2
4. Align the solderless terminals with
the screw holes, and then tighten
the screws.
L:Black
N:White
5. Install the terminal block cover, and fix it with the screws (two).
6. Remove the screws(M4) of the protective ground, align the crimp contacts
with the screw holes, and then tighten the screws.
:Green/ Yellow
The protective ground is internally connected withframe ground.
*
INSTALLATION & CONNECTION
9
9
Chapter
Loosen.
(Turn
counterclockwise
.)
When installing the terminal block to the F350
body, check its orientation. (See the illustration at
the right.)
Attention
Insert side
Right side
2 INSTALLATION & CONNECTION
■Analog input/output connector connection
Connect a strain gauge sensor or voltage output.
Make connections according to the silk-screen printing on the rear panel of the F350.
●Pin Assignment
12SHIELD
2
11-MONITOR(ch2)
10+MONITOR(ch2)
INSTALLATION & CONNECTION
9-MONITOR(ch1)
8+MONITOR(ch1)
7-SIG(ch2)
6+SIG(ch2)
5-SIG(ch1)
4+SIG(ch1)
3-EXC*
2+EXC*
1SHIELD
Compatible connector: ETB42-12P(OSADA CO.,Ltd.) or equivalent
(Connector optional type:CN81)
* EXC is common to ch1 and ch2.
CAUTION
Before connecting a strain gauge sensor, make sure to check the setting of excitation
voltage.
For the setting of excitation voltage, see "4-2 ②Excitation voltage setting" on page
25.
■How to remove the terminal block
1.
Loosen the screws (two) with a screwdriver.
2. Remove the terminal block by giving it a
strong pull.
10
10
2 INSTALLATION & CONNECTION
5mm
Pin No.12
Pin No.1
Adaptable plug ETB42-12P
Tighten.
(Turn clockwise.)
(attachment)
(Turn counterclockwise.)
Loosen.
2KP0Q
2KP0Q
Tighten.
(Turn clockwise.)
Tighten.
(Turn clockwise.)
Sensor cables
F350
A cable is twisted around
a ferrite core.(One roll)
Ferrite core
■How to connect and install the terminal block
Strip 5mm of the covering of the wire to be
1.
connected.
The size of connectable wires is from 0.21 to
3.31mm
2
(AWG12 to 24).
2. Twist the tip to such an extent that it will not spread out.
3. Loosen the screw with a screwdriver to open
the connection hole.
*A Phillips screwdriver 3 to 3.5mm #1 in shaft
diameter is recommended. (Precision screwdriver,
etc.)
4. Insert the wire into the connection hole so as
not to let the tip spread out.
Chapter
2
5. Tighten the screw with the screwdriver.
*0.5Nm of tightening torque is recommended.
6. Lightly pull the wire to make sure that
it is securely clamped.
7. Insert the wire-connected plug into
the F350 body, and tighten the
screws (two).
■Attachment of a ferrite core
It is necessary to twist a power supply cable and sensor cables, such as a load cell, around an
attached ferrite core.
INSTALLATION & CONNECTION
11
11
2 INSTALLATION & CONNECTION
-SIG
-EXC
+EXC
+SIG
SHIELD
+IN
+OUT
-IN
-OUT
5
4
3
2
1
SHIELD
+SIG
-SIG
+IN
+OUT
-IN
-OUT
F350
12
7
6
SHIELD
+SIG
-SIG
12
7
6
-SIG
-EXC
+EXC
+SIG
SHIELD
+
IN
+
OUT
-
IN
-
OUT
F350
+
S
-
S
5
4
3
2
1
+IN
+OUT
-IN
-OUT
+S
-S
Short-circuit +EXC with +S
and -EXC with -S for
connecting a 6-wire straingage sensor.
■Strain gauge sensor connection
■ 4-wire sensor
Chapter
2
INSTALLATION & CONNECTION
■ 6-wire sensor
12
12
2 INSTALLATION & CONNECTION
+
-
External equipment
8 or 10
9 or 11
+ MONITOR1or 2
← F350 Inside Outside →
- MONITOR1or 2
+
-
+
-
+
SIG
-
SIG
A/D
Analog
filter
■Strain gauge sensor cable coloration
Cable colors of sensors may differ from one manufacturer to another (it may even differ from one
model to another for some products). Refer to the sensor manual (or data sheet) and check signal
names and colors in order to connect the cables correctly.
CAUTION
The excitation voltage of the F350 is 2.5V/10V. If the maximum excitation voltage of
the sensor is under 2.5V/10V, heating or damage may result.
■Voltage output connection
The voltage output terminals output analog voltage in proportion to each sensor signal input.
The voltage output is approximately 2V per sensor input of 1mV/V.
It is useful for observing or recording waveforms by connecting a recorder, etc.
・Example) Connection of output equivalent circuit and external instrument.
Chapter
2
・Since the output voltage is taken out in a step prior to A/D conversion of the sensor input signal, it
・Since the ±MONITOR terminals are not insulated from the internal circuit, use two-core shielded
CAUTION
is not synchronized with indicated values processed digitally, such as Digital Zero and digital
filter. Output in synchronization with indicated values requires an optional D/A converter.
twisted pair wires for connection with external equipment, and carry out with as short a wiring as
possible.
● Do not short-circuit. Doing so will cause a failure.
● Do not apply voltage externally. Doing so will cause breakage.
INSTALLATION & CONNECTION
13
13
2 INSTALLATION & CONNECTION
Vceo=30V(max)
Output
Output COM1
Ic=30mA(max)
COM
Input
F350
Sink type
PLC etc.
Plus common connection
A8 ~ A10
B2 ~ B6
A1,A7
B1
DC24V
(Minus common connection)
+24V
0V
A2 ~ A6
COM2
+12V
F350 Inside
IN
IN
・ The external element is required to withstand Ic=10mA.
・ Leakage from the external element is required to be 100μA or below.
B8~12
B7
Open → OFF
Short → ON
Push
Switch
Toggle
SwitchRelay
TTL Open Collector
(ON when IN is HI)
■External I/O connection
Details of signal, see "8-1.External I/O" on page 60.
・How to connect external output (Sink type)
Chapter
2
The external output circuit is operated through an open collector. A1(A7) COM1 is the common
terminal. The open collector output capacity is 30mA and the withstand voltage is up to 30V.
INSTALLATION & CONNECTION
●Output Transistor Status
Output DataTr
・How to connect external input
OFFOFF
ONON
For driving, have an external power source ready.
(24V DC is shown in the illustration.)
(minus common connection of no-voltage contact input type)
A signal is inputted to the signal input circuit by short-circuiting or opening the input terminal and
the COM2 terminal. Short-circuiting is effected by means of a contact (such as a relay or a switch)
or a noncontact (such as a transistor or an open-collector TTL)
14
14
■How to assemble the connector
Connector
Pan-head machine screw
M2×8 (short) (two)
Nut M2 (four)
Pan-head machine screw
M2×10 (long) (two)
Washer (two)
Case (two)
Screw (two)
RNC[
SI/F
A11
← Inside Outside →
F350
A12
Set the connector and screws (two) into the grooves of the case (one side).
1.
2 INSTALLATION & CONNECTION
Chapter
2
2. Cover with the other case, and fit the cases.
3. Tighten the M2×8 pan-head machine screws (two).
Tighten the M2×10 pan-head machine screws (two).
Be aware that washers should be set to the M2×10 pan-head machine screws
(two).
■SI/F interface connection
Two-wire serial interface (SI/F) for connecting printers and external display from UNIPULSE.
Connect from A11 and A12 of the external input/output connector.
The interface is nonpolarized and up to three external devices can be connected.
Wiring materials should use a two-core parallel cable, cabtyre cable (electric wire with covering
thickened for construction), etc. When a two-core parallel cable or a cabtyre cable is used, the
transmission distance is approximately 30m. When a two-core shielded twisted pair wire is used,
the transmission distance is approximately 300m.
Do not parallel it with AC lines and high-voltage lines. Doing so will cause malfunction.
INSTALLATION & CONNECTION
15
2TKPVGT
&KU
%QPXGTVGT
15
3 SETTING MODE CONFIGURATION
3SETTING MODE
CONFIGURATION
Indicated value
display
HI/LO limit setting
mode (P21)
5: A/D conversion speed select
(P35)
1: ch1 Digital filter (P37)
1: Motion Detect (P41)
3: Hysteresis (P45)
2: Near Zero (P47)
1: Comparison function
select (P48、P49)
ESC
F3(Mode3)
Ch Operation
F2(Mode2)
Calculation/common Operation
F1(Mode1)
Comparison/hold
FNC
ESC
HI/LO
H
I
Ch
1
Ch 2
OK
LO
PEAK
HOLD
H
I
LO
H
I
LO
H
I
Ch
1
Ch 2
OK
LO
PEAK
HOLD
H
I
LO
H
I
LO
H
I
Ch
1
Ch 2
OK
LO
PEAK
HOLD
H
I
LO
H
I
LO
2: SI/F function select
(P62)
4: D/Z synchronous mode
(P39)
3: Display frequency
(P43)
7: Average sample number
(P53)
6: Detection time (P57)
5: Calculation hold function
select (P50)
2: ch2 Analog filter
2: ch1 Analog filter (P36)
1: ch2 Digital filter
3: ch2 Zero Tracking
3: ch1 Zero Tracking (P40)
3-1.Setting mode composition
Chapter
3
SETTING MODE CONFIGURATION
16
16
3 SETTING MODE CONFIGURATION
F5(Mode5)
Option
BCODAV/DAI232
2: Communication
condition (P72)
8: Alarm LO limit
(P33)
7: Alarm HI limit
(P33)
6: Digital Zero limit
(P32)
5: ch2 Digital offset
4: ch2 Actual load
calibration
3:ch2 Equivalent input
calibration
2:ch2 Zero calibration
5: ch1 Digital offset
(P31)
4: ch1 Actual load
calibra tion (P30)
3:ch1 Equivalent input
calibration (P28)
1: Calibration function select
(P23)
F9(Mode9)
Protect/initialize
*As for an option setting, a setting item is different depending
F4(Mode4)
Calibration setting
H
I
Ch1Ch 2
OK LO
PEAK
HOLD
HILO
HILO
Ch
1
HILO
H
I
Ch 2
OK LO
PEAK
HOLD
HILO
H
I
Ch1Ch 2
OK LO
PEAK
HOLD
HILO
HILO
2: ch1 Zero calibration
(P27)
Moreover, a set item is not displayed when there is no option.
2: Output rate
(P68)
2: Zero scale value
(P80 )
3: Delimiter
(P72)
3: BCD / binary
select (P68)
3: Full scale value
(P80 )
1: Communication
mode (P71)
1: Output data
select (P66)
1: Output data
select (P79)
4: Password
(P97)
3: Initialize
(P97)
2: Key protect
(P96)
1: Setting protect
(P95)
on the installed option.
Chapter
3
SETTING MODE CONFIGURATION
17
17
Loading...
+ 85 hidden pages
You need points to download manuals.
1 point = 1 manual.
You can buy points or you can get point for every manual you upload.