● The F350 is a digital indicator that is capable of two-channel inputs of strain gauge
sensors.
● Calculations, such as addition and subtraction, and comparison judgments are made
by using the input values of two channels. The F350 is also suitable for balance
measurement.
● The F350 is equipped with comparison judgments on the input value of each channel,
and analog voltage output.
● The A/D conversion speed of each channel and the digital processing speed for
calculations and judgments are as high as 3,000 times/sec. It can also be switched to
300 times/sec.
1 OUTLINE
Chapter
1
OUTLINE
● The F350 is equipped with hold functions with respect to calculated values.
(sample, peak, bottom, and average holds)
● The F350's compact size of DIN96×96 realizes space-saving.
● By mounting an optional interface according to the purpose, ease of load control can
be further improved.
1
1
CAUTION
F350 body・・・1F350 operation manual・・・1
AC input cord・・・1
Conversion plug for
Mini screwdriver for Connector for BCD parallel
(with BCD parallel data output option)
AC input code・・・1
(with D/A converter option)
Ferrite core・・・2
(with unity band)
External input/output
(with solderless terminal)
connector・・・1
terminal block connection・・・1
data output・・・1
1 OUTLINE
Chapter
1
OUTLINE
1-2.Contents of the package
The packaging box contains the following.
Be sure to check them before use.
CAUTION
The attached AC cable is intended for domestic use in Japan, and it is rated as 125V AC, 10A. If it
is used at voltages exceeding the rating or it is used overseas, have a suitable AC cable ready
separately.
2
2
CAUTION
Strain gauge type sensor
(Option)
D/A Converter
Valve, recorder
(Option)
(Option)
PrinterDisplayVarious converters
PC
* +.1* +.1
* +. 11 -
* 1 . &
2 ' # -
%J
%J
(0%'5%
*+.1
<'41
*1.&
F
350
Main status display
Sub status display
Setting key pad
Main numerical display
Sub numerical display
section
section
section
section
1 OUTLINE
1-3.About connectable devices
Chapter
1
OUTLINE
1-4.Appearance description
3
■Front panel
3
CAUTION
In this operation manual, the display status is expressed as follows:
Lighting:Status display section → Black
Numerical display section → White
Blinking:Lighting state + Lines in a radial pattern
Being unlit:Status display section → Gray
Numerical display section → Nothing displayed
LO
H
I
Ch
1
Ch 2
OK
LO
PEAK
HOLD
H
I
LO
H
I
Lighting
Blinking
Being unlit
Chapter
1
OUTLINE
1 OUTLINE
Main status display section
The status with respect to calculated values is indicated.
HI:Lights when the indicated value is larger than the calculation HI limit.
Indicates that the external output "calculation HI" is ON.
OK:Lights when the indicated value is smaller than the calculation HI limit and larger
than the calculation LO limit.
Indicates that the external output "calculation OK" is ON.
LO:Lights when the indicated value is smaller than the calculation LO limit.
Indicates that the external output "calculation LO" is ON.
PEAK:Blinks during hold detection, and lights when a hold is confirmed.
However, it is always unlit in all section sample hold.
HOLD:Lights during hold detection and when a hold is confirmed.
However, it blinks when hold waiting in all section sample hold.
Main numerical display section
The following display patterns exist.
Indicated value:Displays calculation results during measurements.
Error display:Displays an error message when an error occurs.
Setting value: Displays setting values in setting mode.
Sub status display section
The status with respect to ch1/ch2 values is indicated.
HI:Lights when the indicated value is larger than the ch1/ch2 HI limit.
Indicates that the external output "ch1/ch2 HI output" is ON.
LO:Lights when the indicated value is smaller than the ch1/ch2 LO limit.
Indicates that the external output "ch1/ch2 LO output" is ON.
Sub numerical display section
The following display patterns exist.
Indicated value: Displays ch1/ch2 values during measurements.
Error display: Displays an error message when an error occurs.
Setting value:Displays the mode number and item number at the ch1 display section in
setting mode; the currently-selected weight value may be displayed at the ch2
display section according to the setting item
4
4
CAUTION
FNC
ESC
HI/LO
ZERO
HOLD
Setting key pad
These are keys for commanding settings and operations.
・Use this key to go to a setting mode state.
・Use this key to cancel setting/execution.
・Use this key to go back from a setting mode state to an indication display state.
・Use this key to take a shortcut from an indication display state to HI/LO limit setting.
・
Use this key to go from an indication display state to ch1 or ch2 Digital Zero ready.
・Use this key to execute/release the calculated value hold in an indication display state.
・Use this key to switch between ch1 setting and ch2 setting, etc., in a Digital Zero
ready state or setting item select state.
・Use this key to set whether or not to sign a setting value, when available, in a setting
value input state.
1 OUTLINE
Chapter
1
OUTLINE
・Use this key to select a trailing setting item number in a setting mode state.
・Use this key to select the previous item number in a setting item select state.
・Use this key to increment the numerical value by one in a setting value input state.
・Use this key to select a leading setting item number in a setting mode state.
・Use this key to select the next item number in a setting item select state.
・Use this key to decrement the numerical value by one in a setting value input state.
・Use this key to execute Digital Zero reset in a Digital Zero ready state.
・Use this key to increment the leading setting mode number by one in a setting mode
state.
・Use this key to move the setting digit by one in the lower direction in a setting value
input state.
・Use this key to execute Digital Zero in a Digital Zero ready state.
・Use this key to move to a setting value input state in a setting item select state.
・Use this key to validate the setting value in a setting value input state.
5
5
CAUTION
Options slot
Power input terminal
Protective ground
Signal input/output
Analog input/output
connector
connector
Frame ground
Chapter
1
OUTLINE
1 OUTLINE
■Rear panel
Analog input/output connector
Connect a strain gauge sensor or voltage output.
* For connection, see "
■ Analog input/output connector connection" on page 10.
Signal input/output connector
Connect external input/output signals or SI/F (serial interface).
* For connection, see "
■ External I/O connection" on page 14.
Power input terminal block
Connect an AC input cord.
The input power source is free in the range of 100 to 240V AC (-15%, +10%). The frequency is 50/
60Hz.
* For connection, see "
■ Power input terminal connection" on page 9.
Protective ground
Be sure to ground the protective ground terminal to prevent electric shocks.
Frame ground
6
6
Please ground the frame ground terminal to prevent failures due to static electricity.
(The frame and the frame ground terminal are conducted.)
It may be better to remove depending on the environment of the installation location.
Options slot
One option board can in stall in the option slot.
・BCD parallel data output(BCO)
・RS-232C interface(232)
・D/A converter voltage output(DAV)
・D/A converter current output (DAI)
2INSTALLATION & CONNECTION
2I
NSTALLATION & CONNECTION
The following are precautions related to connection to the signal input/output terminal block.
The precautions described here are important for safety.
Make connections after properly understanding the description.
WARNING
● Do not connect commercial power directly to the signal input/output terminals.
● Make connection to the signal input/output terminals with no power applied
because you may receive an electric shock.
● For connections to the signal input/output terminals, carry out wiring properly after
confirming the signal names and pin assignment numbers.
● Use of the F350 is limited to category II specified by EN61010. Overvoltage applied
to the signal input/output terminals should not exceed the value defined to category
II.
● After wiring, make sure to mount the attached terminal block cover. Otherwise, you
may receive an electric shock.
● Turn on the power after carefully checking the wiring, etc.
Chapter
2
INSTALLATION & CONNECTION
● Do not touch the signal input/output terminals while the power is on. Otherwise, you
may receive an electric shock, or a malfunction may occur.
● The F350 conforms with EMC Directives as a product used in industrial
environments (class A). If it is used in living environments, radio disturbance may
occur. In that case, take suitable measures.
CAUTION
● Tighten the terminal screws at specified torque. If the terminal screws are loose,
short-circuiting, fire, or a malfunction may occur.
● Use shielded cables.
7
7
2 INSTALLATION & CONNECTION
92mm
Panel-cut
+1
0
–
Panel thickness
1.6 to 3.2mm
dimensions
92mm
+1
0
–
2-1.Installation
To install the F350 into a control panel, use the following procedure.
1. Make a hole in the panel according to
Chapter
2
INSTALLATION & CONNECTION
the panel-cut dimensions.
2. Remove the screws (two), and remove
the guide rails from both sides.
* Do not use other screws than those installed to
the F350 body.
3. Fit in the F350 from the front of the panel.
4. Install from the rear the guide rails having
been removed from both sides in Step 2,
and fix them with the screws (two).
8
8
2-2.Connection
Within 6mm
Black
White
Green/Yellow
Color description are for the attached
AC input cord.
■Power input terminal connection
Connect an AC input cord. The input power source is free in the range of 100 to 240V AC (-15%,
+10%).
The frequency is 50/60Hz.
Make connections to the terminal block with a solderless terminal within 6mm in diameter as
shown in the illustration so as not to let the tip of the cable spread out.
1. Make sure that no power is applied.
2. Remove the screws (two), and remove
the terminal block cover.
3. Remove the two screws(M3) at
the terminal block
.
2 INSTALLATION & CONNECTION
Chapter
2
4. Align the solderless terminals with
the screw holes, and then tighten
the screws.
L:Black
N:White
5. Install the terminal block cover, and fix it with the screws (two).
6. Remove the screws(M4) of the protective ground, align the crimp contacts
with the screw holes, and then tighten the screws.
:Green/ Yellow
The protective ground is internally connected withframe ground.
*
INSTALLATION & CONNECTION
9
9
Chapter
Loosen.
(Turn
counterclockwise
.)
When installing the terminal block to the F350
body, check its orientation. (See the illustration at
the right.)
Attention
Insert side
Right side
2 INSTALLATION & CONNECTION
■Analog input/output connector connection
Connect a strain gauge sensor or voltage output.
Make connections according to the silk-screen printing on the rear panel of the F350.
●Pin Assignment
12SHIELD
2
11-MONITOR(ch2)
10+MONITOR(ch2)
INSTALLATION & CONNECTION
9-MONITOR(ch1)
8+MONITOR(ch1)
7-SIG(ch2)
6+SIG(ch2)
5-SIG(ch1)
4+SIG(ch1)
3-EXC*
2+EXC*
1SHIELD
Compatible connector: ETB42-12P(OSADA CO.,Ltd.) or equivalent
(Connector optional type:CN81)
* EXC is common to ch1 and ch2.
CAUTION
Before connecting a strain gauge sensor, make sure to check the setting of excitation
voltage.
For the setting of excitation voltage, see "4-2 ②Excitation voltage setting" on page
25.
■How to remove the terminal block
1.
Loosen the screws (two) with a screwdriver.
2. Remove the terminal block by giving it a
strong pull.
10
10
2 INSTALLATION & CONNECTION
5mm
Pin No.12
Pin No.1
Adaptable plug ETB42-12P
Tighten.
(Turn clockwise.)
(attachment)
(Turn counterclockwise.)
Loosen.
2KP0Q
2KP0Q
Tighten.
(Turn clockwise.)
Tighten.
(Turn clockwise.)
Sensor cables
F350
A cable is twisted around
a ferrite core.(One roll)
Ferrite core
■How to connect and install the terminal block
Strip 5mm of the covering of the wire to be
1.
connected.
The size of connectable wires is from 0.21 to
3.31mm
2
(AWG12 to 24).
2. Twist the tip to such an extent that it will not spread out.
3. Loosen the screw with a screwdriver to open
the connection hole.
*A Phillips screwdriver 3 to 3.5mm #1 in shaft
diameter is recommended. (Precision screwdriver,
etc.)
4. Insert the wire into the connection hole so as
not to let the tip spread out.
Chapter
2
5. Tighten the screw with the screwdriver.
*0.5Nm of tightening torque is recommended.
6. Lightly pull the wire to make sure that
it is securely clamped.
7. Insert the wire-connected plug into
the F350 body, and tighten the
screws (two).
■Attachment of a ferrite core
It is necessary to twist a power supply cable and sensor cables, such as a load cell, around an
attached ferrite core.
INSTALLATION & CONNECTION
11
11
2 INSTALLATION & CONNECTION
-SIG
-EXC
+EXC
+SIG
SHIELD
+IN
+OUT
-IN
-OUT
5
4
3
2
1
SHIELD
+SIG
-SIG
+IN
+OUT
-IN
-OUT
F350
12
7
6
SHIELD
+SIG
-SIG
12
7
6
-SIG
-EXC
+EXC
+SIG
SHIELD
+
IN
+
OUT
-
IN
-
OUT
F350
+
S
-
S
5
4
3
2
1
+IN
+OUT
-IN
-OUT
+S
-S
Short-circuit +EXC with +S
and -EXC with -S for
connecting a 6-wire straingage sensor.
■Strain gauge sensor connection
■ 4-wire sensor
Chapter
2
INSTALLATION & CONNECTION
■ 6-wire sensor
12
12
2 INSTALLATION & CONNECTION
+
-
External equipment
8 or 10
9 or 11
+ MONITOR1or 2
← F350 Inside Outside →
- MONITOR1or 2
+
-
+
-
+
SIG
-
SIG
A/D
Analog
filter
■Strain gauge sensor cable coloration
Cable colors of sensors may differ from one manufacturer to another (it may even differ from one
model to another for some products). Refer to the sensor manual (or data sheet) and check signal
names and colors in order to connect the cables correctly.
CAUTION
The excitation voltage of the F350 is 2.5V/10V. If the maximum excitation voltage of
the sensor is under 2.5V/10V, heating or damage may result.
■Voltage output connection
The voltage output terminals output analog voltage in proportion to each sensor signal input.
The voltage output is approximately 2V per sensor input of 1mV/V.
It is useful for observing or recording waveforms by connecting a recorder, etc.
・Example) Connection of output equivalent circuit and external instrument.
Chapter
2
・Since the output voltage is taken out in a step prior to A/D conversion of the sensor input signal, it
・Since the ±MONITOR terminals are not insulated from the internal circuit, use two-core shielded
CAUTION
is not synchronized with indicated values processed digitally, such as Digital Zero and digital
filter. Output in synchronization with indicated values requires an optional D/A converter.
twisted pair wires for connection with external equipment, and carry out with as short a wiring as
possible.
● Do not short-circuit. Doing so will cause a failure.
● Do not apply voltage externally. Doing so will cause breakage.
INSTALLATION & CONNECTION
13
13
2 INSTALLATION & CONNECTION
Vceo=30V(max)
Output
Output COM1
Ic=30mA(max)
COM
Input
F350
Sink type
PLC etc.
Plus common connection
A8 ~ A10
B2 ~ B6
A1,A7
B1
DC24V
(Minus common connection)
+24V
0V
A2 ~ A6
COM2
+12V
F350 Inside
IN
IN
・ The external element is required to withstand Ic=10mA.
・ Leakage from the external element is required to be 100μA or below.
B8~12
B7
Open → OFF
Short → ON
Push
Switch
Toggle
SwitchRelay
TTL Open Collector
(ON when IN is HI)
■External I/O connection
Details of signal, see "8-1.External I/O" on page 60.
・How to connect external output (Sink type)
Chapter
2
The external output circuit is operated through an open collector. A1(A7) COM1 is the common
terminal. The open collector output capacity is 30mA and the withstand voltage is up to 30V.
INSTALLATION & CONNECTION
●Output Transistor Status
Output DataTr
・How to connect external input
OFFOFF
ONON
For driving, have an external power source ready.
(24V DC is shown in the illustration.)
(minus common connection of no-voltage contact input type)
A signal is inputted to the signal input circuit by short-circuiting or opening the input terminal and
the COM2 terminal. Short-circuiting is effected by means of a contact (such as a relay or a switch)
or a noncontact (such as a transistor or an open-collector TTL)
14
14
■How to assemble the connector
Connector
Pan-head machine screw
M2×8 (short) (two)
Nut M2 (four)
Pan-head machine screw
M2×10 (long) (two)
Washer (two)
Case (two)
Screw (two)
RNC[
SI/F
A11
← Inside Outside →
F350
A12
Set the connector and screws (two) into the grooves of the case (one side).
1.
2 INSTALLATION & CONNECTION
Chapter
2
2. Cover with the other case, and fit the cases.
3. Tighten the M2×8 pan-head machine screws (two).
Tighten the M2×10 pan-head machine screws (two).
Be aware that washers should be set to the M2×10 pan-head machine screws
(two).
■SI/F interface connection
Two-wire serial interface (SI/F) for connecting printers and external display from UNIPULSE.
Connect from A11 and A12 of the external input/output connector.
The interface is nonpolarized and up to three external devices can be connected.
Wiring materials should use a two-core parallel cable, cabtyre cable (electric wire with covering
thickened for construction), etc. When a two-core parallel cable or a cabtyre cable is used, the
transmission distance is approximately 30m. When a two-core shielded twisted pair wire is used,
the transmission distance is approximately 300m.
Do not parallel it with AC lines and high-voltage lines. Doing so will cause malfunction.
INSTALLATION & CONNECTION
15
2TKPVGT
&KU
%QPXGTVGT
15
3 SETTING MODE CONFIGURATION
3SETTING MODE
CONFIGURATION
Indicated value
display
HI/LO limit setting
mode (P21)
5: A/D conversion speed select
(P35)
1: ch1 Digital filter (P37)
1: Motion Detect (P41)
3: Hysteresis (P45)
2: Near Zero (P47)
1: Comparison function
select (P48、P49)
ESC
F3(Mode3)
Ch Operation
F2(Mode2)
Calculation/common Operation
F1(Mode1)
Comparison/hold
FNC
ESC
HI/LO
H
I
Ch
1
Ch 2
OK
LO
PEAK
HOLD
H
I
LO
H
I
LO
H
I
Ch
1
Ch 2
OK
LO
PEAK
HOLD
H
I
LO
H
I
LO
H
I
Ch
1
Ch 2
OK
LO
PEAK
HOLD
H
I
LO
H
I
LO
2: SI/F function select
(P62)
4: D/Z synchronous mode
(P39)
3: Display frequency
(P43)
7: Average sample number
(P53)
6: Detection time (P57)
5: Calculation hold function
select (P50)
2: ch2 Analog filter
2: ch1 Analog filter (P36)
1: ch2 Digital filter
3: ch2 Zero Tracking
3: ch1 Zero Tracking (P40)
3-1.Setting mode composition
Chapter
3
SETTING MODE CONFIGURATION
16
16
3 SETTING MODE CONFIGURATION
F5(Mode5)
Option
BCODAV/DAI232
2: Communication
condition (P72)
8: Alarm LO limit
(P33)
7: Alarm HI limit
(P33)
6: Digital Zero limit
(P32)
5: ch2 Digital offset
4: ch2 Actual load
calibration
3:ch2 Equivalent input
calibration
2:ch2 Zero calibration
5: ch1 Digital offset
(P31)
4: ch1 Actual load
calibra tion (P30)
3:ch1 Equivalent input
calibration (P28)
1: Calibration function select
(P23)
F9(Mode9)
Protect/initialize
*As for an option setting, a setting item is different depending
F4(Mode4)
Calibration setting
H
I
Ch1Ch 2
OK LO
PEAK
HOLD
HILO
HILO
Ch
1
HILO
H
I
Ch 2
OK LO
PEAK
HOLD
HILO
H
I
Ch1Ch 2
OK LO
PEAK
HOLD
HILO
HILO
2: ch1 Zero calibration
(P27)
Moreover, a set item is not displayed when there is no option.
2: Output rate
(P68)
2: Zero scale value
(P80 )
3: Delimiter
(P72)
3: BCD / binary
select (P68)
3: Full scale value
(P80 )
1: Communication
mode (P71)
1: Output data
select (P66)
1: Output data
select (P79)
4: Password
(P97)
3: Initialize
(P97)
2: Key protect
(P96)
1: Setting protect
(P95)
on the installed option.
Chapter
3
SETTING MODE CONFIGURATION
17
17
3 SETTING MODE CONFIGURATION
Ch
1
H
I
LO
Ch
1
H
I
LO
Selecting a
Select a setting item.
The highest digit blinks to enter setting.
Select the digit you want to set.
Set a setting value.
The setting value is validated.
Go back to setting mode selection.
From the indicated value display,
Selecting a
Selecting a setting
"F2" of setting mode 2
For signing the setting value, when
available, press.
"F9" of setting mode 9
FNC
ESC
Go back to setting mode 1.
・・・
ESC
Go back to the indicated value.
Ch
1
HILO
Ch
1
H
I
LO
For setting values available for ch1 and ch2 individually,
and inputting a
setting value
is displayed. is displayed.
setting mode.
setting item.
"F1" of setting mode 1 is displayed
on the SUB display (ch1 side).
switch with .
If key operation is not performed for one minute or more between the time of pressing
the key and the time of validating with the key, the setting is stopped to go
back to the indicated value display. In that case, the setting value is not saved.
The protect state of each setting mode is visible at a glance.
Protect information
Protect ON
Protect OFF
For the protect setting method, see " ■ Setting protect setting
method " on page 95 .
3 SETTING MODE CONFIGURATION
Chapter
3
SETTING MODE CONFIGURATION
19
19
3 SETTING MODE CONFIGURATION
H
I
Ch
1
Ch 2
OK
LO
PEAK
HOLD
H
I
LO
H
I
LO
H
I
Ch
1
Ch 2
OK
LO
PEAK
HOLD
H
I
LO
H
I
LO
【Normal setting item】
【Normal setting item (with separate
H
I
Ch
1
Ch 2
OK
LO
PEAK
HOLD
H
I
LO
H
I
LO
Selected ch display
1:ch1
2:ch2
Item number (1 - 9)
Mode number (1 - 5, 9)
【Calibration operation item】
(Zero calibration, equivalent input
Selected ch display
Example) 1 → ch1 selected
Indicated value
Example) ch1 indicated value 99.97
settings on ch1 and ch2)】
calibration, actual load calibration)
H
I
Ch
1
Ch 2
OK
LO
PEAK
HOLD
H
I
LO
H
I
LO
H
I
Ch
1
Ch 2
OK
LO
PEAK
HOLD
H
I
LO
H
I
LO
Ch1 setting value Ch2 setting value
Chapter
3
SETTING MODE CONFIGURATION
≪Setting item select state≫
≪Setting ch switching function≫
For settings available for ch1 and ch2 in a setting item select state, press to switch
between ch1 value setting and ch2 value setting.
20
20
3 SETTING MODE CONFIGURATION
HI/LO
Ch 2
H
I
LO
Ch 2
H
I
LO
LO
Ch 2
H
I
H
I
Ch 2
LO
Ch 2
H
I
LO
Ch 2
H
I
LO
HI/LO
H
I
Ch
1
Ch 2
OK
LO
PEAK
HOLD
H
I
LO
HILO
H
I
Ch
1
OK
LO
PEAK
HOLD
H
I
LO
H
I
Ch
1
OK
LO
PEAK
HOLD
H
I
LO
H
I
Ch
1
OK
LO
PEAK
HOLD
H
I
LO
H
I
Ch
1
OK
LO
PEAK
HOLD
H
I
LO
H
I
Ch
1
OK
LO
PEAK
HOLD
H
I
LO
H
I
Ch
1
OK
LO
PEAK
HOLD
H
I
LO
HI/LO
HI/LO
ESC
Indicated value display
ch1 LO limit
ch1 HI limit
ch2 LO limit
ch2 HI limit
Calculation LO limit
Calculation
Exit the HI/LO limit setting item display
HI/LO
HI limit
state.
≪HI/LO limit setting mode function≫
By pressing on the indicated value display screen, you can enter a HI/LO limit setting
item select state.
Press at the setting value you want to change, and you will go to a setting value input
state.
After inputting the setting value, you will go back to the HI/LO limit setting item state.
Chapter
3
SETTING MODE CONFIGURATION
21
21
3 SETTING MODE CONFIGURATION
H
I
Ch
1
Ch 2
OK
LO
PEAK
HOLD
H
I
LO
H
I
LO
FNC
H
I
Ch
1
Ch 2
OK
LO
PEAK
HOLD
H
I
LO
H
I
LO
H
I
Ch
1
Ch 2
OK
LO
PEAK
HOLD
H
I
LO
H
I
LO
H
I
Ch
1
Ch 2
OK
LO
PEAK
HOLD
H
I
LO
H
I
LO
H
I
Ch
1
Ch 2
OK
LO
PEAK
HOLD
H
I
LO
H
I
LO
ESC
H
I
Ch
1
Ch 2
OK
LO
PEAK
HOLD
H
I
LO
HILO
While the lowest decimal place on the sub display
(ch2 side) is lit, writing in NOVRAM is in progress.
If the power is turned off at this time, normal writing
cannot be performed.
H
I
Ch
1
Ch 2
OK
LO
PEAK
HOLD
H
I
LO
HILO
3-3.Calling a setting mode
Example) For changing the ch1 digital filter from "000" to "256"
1. Press from an indicated value display state.
Chapter
3
SETTING MODE CONFIGURATION
2. Go to a setting mode select state.
Press twice to make the mode number "3."
3. Go to a setting item select state.
Press once to select F3-1.
Press .
4. Go to a setting value input state.
Press to move the setting digit.
Press to change the numerical value.
Press to validate it.
The change can be cancelled by pressing .
While a signed setting value is being input, press to select whether or not to sign it.
5. Upon validation, you will go back to the setting item
select state.
22
22
4-1.How to calibration
4CALIBRATION PROCEDURE
A data sheet will be attached to a strain gauge sensor you buy.
On the data seat, the following values are listed.
CapacityRated Output
Non-LinearityHysterisis
Input ResistanceOutput Resistance
Zero Balance etc.
The Capacity and the Rated Output are necessary values for the equivalent input
calibration. Input these two values to F350.
POINT
"Calibration" refers to an operation whereby matching between the F350 and a strain gauge sensor
is obtained. The F350 uses the two calibration methods as described below.
■Equivalent input calibration
Calibration is performed without an actual load by entering the rated output value (mV/V) and the
capacity (to be indicated) of the strain-gage sensor by the keys. Calibration is easily performed
when no actual load is available.
For example,
for load:2.001mV/V-100.0kgf
for pressure:2.002mV/V-10.00kgf/cm
for torque:2.502mV/V-15.00kgf・m
4 CALIBRATION PROCEDURE
Chapter
4
2
Gain will be automatically decided by registering the values indicated as follows.
■Actual load calibration
This approach provides calibration by applying an actual load to the strain gauge sensor and
inputting the actual load value. This calibration is without little errors and more correct.
CALIBRATION PROCEDURE
Perform either actual load calibration or equivalent input calibration.
23
*
If an actual load cannot be applied, make sure to perform equivalent input calibration
alone.
23
Equivalent input
Actual load
calibration
calibration
Digital offset
⑦
Calibration protect
Min. scale division
a.Equivalentinput
Calibration protect
①
④
⑥
⑩
Zero calibration
⑤
Excitation voltage
②
Decimal place
③
Digital Zero limit
⑧
b.Actual load
⑥
Alarm HI and LO limits
⑨
calibration
calibration
Chapter
4
CALIBRATION PROCEDURE
4 CALIBRATION PROCEDURE
4-2.Procedure of equivalent input calibration and
calibration protect
The equivalent input calibration and the calibration protect uses the following procedure :
Turn off the calibration protect that inhibits
calibration.
Set the excitation voltage according to the sensor
used.
Set the decimal place.
Set the desired minimum value of digital increments.
(This step may be omitted if there is no change.)
Set the zero point of the strain gauge sensor in no-
load condition (with the sensor unloaded).
Equivalent input calibration
Register the rated output value of the strain gauge
sensor, and the indicated value of that time.
Actual load calibration
Set the span (gain) point of the strain gauge sensor
with a load applied to the sensor.
The calibrated value can be offset in advance.
(When not using, set "0.")
Set the load limit that allows Digital Zero.
24
Set the HI/LO limit for monitoring overload.
Turn on the calibration protect to prevent
misoperation.
24
①Releasing the calibration protect
Ch
1
Ch 2
H
I
LO
H
I
LO
FNC
H
I
Ch
1
Ch 2
OK
LO
PEAK
HOLD
H
I
LO
H
I
LO
0 0 0
Calibration protect
0:OFF 1:ON
System protect
0:OFF 1:ON
Work protect
0:OFF 1:ON
《Setting value》
ESC
Turn off the calibration protect that inhibits calibration.
≪Setting value≫
0:OFFEnables rewriting of setting values.
1:ONDisables rewriting of setting values.
◇Calibration protect setting method
1. Select setting mode 9.
Press → five times.
4 CALIBRATION PROCEDURE
2. Set "Calibration protect."
Press once. ⇒ "Setting protect" is displayed.
Press → twice to make the digit of the
calibration protect blink.
Input the setting value with .
(Initial value: 0)
Press to validate it.
3. Press twice to go back to the indicated value
display.
②Excitation voltage setting
Select the bridge voltage (excitation voltage) supplied to the strain gauge sensor.
The bridge voltage can be selected from 2.5V and 10V.
After this setting, make sure to perform calibration.
≪Setting value≫
0:2.5V
1:10V
Chapter
4
CALIBRATION PROCEDURE
25
* Common to ch1 and ch2.
③Decimal place setting
Set the decimal place for load-value-related display, setting items, etc.
≪Setting value≫
0:None
1:0.0
2:0.00
3:0.000
* Calculated values are also synchronized with this setting.
25
4 CALIBRATION PROCEDURE
Ch
1
Ch 2
H
I
LO
H
I
LO
FNC
H
I
Ch
1
Ch 2
OK
LO
PEAK
HOLD
H
I
LO
H
I
LO
002
Min. scale division
Excitation voltage
0:2.5V1:10V
Decimal place
0:None
1:0.0
2:0.00
3:0.000
0:1 4:20
1:2 5:50
2:5 6:100
3:10
《Setting value》
ESC
The maximum excitation voltage value of the strain gauge sensor connected to the
F350 should always be larger than the set bridge voltage.
If the excitation voltage of the F350 is larger than the maximum excitation voltage
value of the strain gauge sensor, the strain gauge sensor may heat or break.
④Min. scale division setting
Set the minimum unit (scale interval, scale division) of measurements.
CAUTION
Chapter
4
CALIBRATION PROCEDURE
≪Setting value≫
0:1
1:2
2:5
3:10
4:20
5:50
6:100
* Calculated values are also synchronized with this setting
◇Excitation voltage, decimal place, and min. scale division setting method
1. Select setting mode 4.
Press → three times.
2. Set "Excitation voltage," "Decimal place," and "Min.
scale division."
Press once. ⇒ "Calibration function select"
is displayed.
Press , and the highest digit blinks.
26
26
① Select excitation voltage.
Input the excitation voltage with, and
move to the lower digit with. (Initial value: 0)
② Select decimal place.
Input the decimal place with, and
move to the lower digit with. (Initial value: 2)
③ Select min. scale division.
Input the min. scale division with, and
validate it with.(Initial value: 0).
3. Presstwice to go back to the indicted value display.
4 CALIBRATION PROCEDURE
Ch
1
Ch 2
H
I
LO
H
I
LO
FNC
H
I
Ch
1
Ch 2
OK
LO
PEAK
HOLD
H
I
LO
HILO
Selected ch display
H
I
Ch
1
Ch 2
OK
LO
PEAK
HOLD
H
I
LO
H
I
LO
H
I
Ch
1
Ch 2
OK
LO
PEAK
HOLD
HILO
HILO
ESC
⑤Zero calibration
Register the zero point with the sensor unloaded. Also, if there is any initial load, such as a jig, for
measurements, the zero point can be registered with the initial load applied, but the signal input
range needs to be considered as the initial load is subtracted from it.
In addition, there is a way of considering the initial load by using the digital offset. See " ⑦Digital
offset" on page 31.
≪Setting value≫
-3.000~+3.000mV/V
By performing zero calibration, the setting value of zero calibration becomes the sensor output
value (mV/V) of the time when it was executed. The load with which actual load calibration was
performed can be confirmed. However, it stays unchanged when an error occurs.
Chapter
◇Zero calibration setting method (Example: ch1 zero calibration)
1. Select setting mode 4.
Press → three times.
2. Set "Zero calibration."
Press twice. ⇒ "Zero calibration" is displayed.
* For performing ch2 zero calibration, switch with
.
Enter a setting value input state with , and the entire
setting value blinks. (Initial value: 0.000)
Unload the sensor, check the current load value on the sub
display (ch2), and press .
3. Zero calibration is executed.
A message as shown on the right-hand side is displayed.
If zero calibration is not executed normally, the alarm
sounds and an error is displayed.
4
CALIBRATION PROCEDURE
27
4. Go back to the setting item select state.
If zero calibration is executed normally, you will go back
to the setting item select state, and the output (mV/V) of
the time when it was performed will appear.
The example shows that zero calibration was performed at
0.017mV/V.
5. Press twice to go back to the indicated value display.
27
Chapter
Ch
1
Ch 2
H
I
LO
H
I
LO
FNC
H
I
Ch
1
Ch 2
OK
LO
PEAK
HOLD
HILO
H
I
LO
Selected ch display
H
I
Ch
1
Ch 2
OK
LO
PEAK
HOLD
H
I
LO
H
I
LO
4CALIBRATION PROCEDURE
⑥a.Equivalent input calibration
See the test report of the sensor, and register the rated output value and the value you want to
display at that time (rated capacity value).
Input the rated output value and display value (rated capacity value) successively.
Equivalent input calibration is not executed by simply inputting the rated output value.
≪Setting value≫
Rated output value [mV/V]:-3.000 ~ 3.000 (except 0)Display value (rated capacity): -19999 ~ 19999 (except 0)
By performing equivalent input calibration, the setting value of actual load calibration becomes the
input rated capacity value.
However, it stays unchanged when an error occurs.
4
CALIBRATION PROCEDURE
◇Equivalent input calibration setting method
(Example: ch1 equivalent input calibration)
1. Select setting mode 4.
Press→three times.
2. Set "Rated output value" of the sensor.
Press three times. ⇒ "Equivalent input
calibration" is displayed.
* For performing ch2 equivalent input calibration,
switch with .
Enter a setting value input state with , and the
highest digit blinks. (Initial value: 3.000)
Check the test report of the sensor in use, and input the
rated output value (mV/V) with .
Move to the lower digits with .
28
28
Press to validate it.
3. Set "Rated capacity value."
Decide the load you want to display at the rated output
value, and input it with .
Move to the lower digits with .
Press to validate it.
* For inputting a negative sign, press
.
4. Equivalent input calibration is executed.
H
I
Ch
1
Ch 2
OK
LO
PEAK
HOLD
HILO
H
I
LO
H
I
Ch
1
Ch 2
OK
LO
PEAK
HOLD
HILO
H
I
LO
H
I
Ch
1
Ch 2
OK
LO
PEAK
HOLD
H
I
LO
H
I
LO
If the calibration error()is displayed, take measures according to each
error, and redo calibration. For details, see "11-7 Error/message list" on page 97.
After completion of calibration, make sure to turn on the calibration protect.
Point
A message as shown on the right-hand side is displayed.
If equivalent input calibration is not executed normally,
the alarm sounds.
5. If equivalent input calibration is executed normally,
you will go back to the setting item select state.
4 CALIBRATION PROCEDURE
Chapter
4
CALIBRATION PROCEDURE
29
29
4 CALIBRATION PROCEDURE
Ch
1
Ch 2
H
I
LO
H
I
LO
FNC
H
I
Ch
1
Ch 2
OK
LO
PEAK
HOLD
HILO
H
I
LO
H
I
Ch
1
Ch 2
OK
LO
PEAK
HOLD
H
I
LO
H
I
LO
Selected ch display
H
I
Ch
1
Ch 2
OK
LO
PEAK
HOLD
H
I
LO
H
I
LO
H
I
Ch
1
Ch 2
OK
LO
PEAK
HOLD
H
I
LO
HILO
⑥b.Actual load calibration
Apply an actual load, and register the value of the actual load (rated capacity value).
≪Setting value≫
-19999~19999 (except 0)
By performing actual load calibration, the setting value of equivalent input calibration becomes the
sensor output value (mV/V) of the time when it was executed. The load with which actual load
calibration was performed can be confirmed. However, it stays unchanged when an error occurs.
◇Actual load calibration setting method
Chapter
4
CALIBRATION PROCEDURE
(Example: ch1 actual load calibration)
1. Select setting mode 4.
Press → three times.
2. Set "Actual load" of the sensor.
Press four times. ⇒ "Actual load calibration"
is displayed.
* For performing actual load calibration on ch2,
switch with .
Enter a setting value input state with, and the
highest digit blinks. (Initial value: 100.00)
Check the capacity of the actually-applied load, and
input the value with .
* For inputting a negative sign, press
Move to the lower digits with .
Recheck that the load is properly applied on the sub display (ch2 side), and press
to validate it.
.
30
30
3. Actual load calibration is executed.
If actual load calibration is executed normally, a
message as shown on the right-hand side is displayed.
If actual load calibration is not executed normally, the
alarm sounds and an error is displayed.
4. If actual load calibration is executed normally, you will
go back to the setting item select state.
4 CALIBRATION PROCEDURE
Ch
1
Ch 2
H
I
LO
HILO
FNC
H
I
Ch
1
Ch 2
OK
LO
PEAK
HOLD
HILO
H
I
LO
Selected ch display
ESC
⑦Digital offset
This function is to subtract a set value from the indicated value. This function is convenient when
zero cannot be obtained with no load for some reason or for offsetting. When not using, set "0."
(Displayed value)=(Actual indicated value)-(Setting value of digital offset)
≪Setting value≫
-19999~19999
◇Digital offset setting method
1. Select setting mode 4.
Press→three times.
2. Set "Digital offset."
Press five times. ⇒ "Digital offset" is displayed.
* Switch desired ch with
.
Press , and the highest digit blinks.
Input the setting value with .
Move to the lower digits with .
Validate it with .
* For inputting a negative sign, press
.
3. Press twice to go back to the indicated value display.
Chapter
4
CALIBRATION PROCEDURE
31
31
Chapter
Ch
1
Ch 2
H
I
LO
HILO
FNC
H
I
Ch
1
Ch 2
OK
LO
PEAK
HOLD
H
I
LO
H
I
LO
ESC
4CALIBRATION PROCEDURE
⑧Digital Zero limit
Set the load limit that allows Digital Zero.
If the absolute value of the load at which Digital Zero is about to be performed is larger than the
setting value, an error will result, and the indicated value will not be zeroed.
≪Setting value≫
0~19999
* The setting value is common to ch1 and ch2.
The amounts of Digital Zero obtained from the point where zero calibration was performed are
monitored.
4
CALIBRATION PROCEDURE
◇Digital Zero limit setting method
1. Select setting mode 4.
Press→three times.
2. Set "Digital Zero limit."
Press six times. ⇒ "Digital Zero limit" is
displayed.
Press , and the highest digit blinks.
Input the setting value with.
Move to the lower digits with .
Validate it with .
3. Press twice to go back to the indicated value display.
32
32
4 CALIBRATION PROCEDURE
Ch
1
Ch 2
H
I
LO
H
I
LO
FNC
H
I
Ch
1
Ch 2
OK
LO
PEAK
HOLD
H
I
LO
H
I
LO
H
I
Ch
1
Ch 2
OK
LO
PEAK
HOLD
HILO
H
I
LO
ESC
⑨Alarm HI and LO limits
Overload of the ch1 value and ch2 value is monitored.
When they exceed the alarm HI limit or fall below the alarm LO limit, the external output "ALM"
turns ON.
Alarm HI limit:-19999 ~ 19999 (invalid at 19999 (initial value))
Alarm LO limit:-19999 ~ 19999 (invalid at -19999 (initial value))
* The setting value is common to ch1 and ch2.
・ For initial value, since both the alarm HI and LO limits are invalid as default settings, set them
again for use.
・Since the alarm HI limit and LO limit are independent, simple use of either one of them is also
possible.
◇Alarm HI and LO limits setting method
Chapter
4
1. Select setting mode 4.
Press→three times.
2. Set "Alarm HI limit."
Press seven times. ⇒ "Alarm HI limit" is
displayed.
Press , and the highest digit blinks.
Input the setting value with .
Move to the lower digits with .
Validate it with . (Initial value: 99999)
* For inputting a negative sign, press while the input digit is blinking.
3. Set "Alarm LO limit."
Press once. ⇒ "Alarm LO limit" is displayed.
Press , and the highest digit blinks.
Input the setting value with .
CALIBRATION PROCEDURE
Move to the lower digits with .
Validate it with . (Initial value: -99999)
* For inputting a negative sign, press while the input digit is blinking.
4. Presstwice to go back to the indicated value display.
33
33
4 CALIBRATION PROCEDURE
Ch
1
Ch 2
H
I
LO
HILO
FNC
H
I
Ch
1
Ch 2
OK
LO
PEAK
HOLD
H
I
LO
H
I
LO
0 0 0
Calibration protect
0:OFF 1:ON
System protect
0:OFF 1:ON
Work protect
0:OFF 1:ON
《Setting value》
ESC
⑩Locking the calibration protect
Turn on the calibration protect that inhibits calibration.
≪Setting value≫
0:OFF Enables rewriting of setting values.
1:ONDisables rewriting of setting values.
◇Calibration protect setting method
1. Select setting mode 9.
Press →five times.
Chapter
4
CALIBRATION PROCEDURE
2. Set "Calibration protect."
Pressonce. ⇒ "Setting protect" is displayed.
Press→twice to make the digit of the
calibration protect blink.
Input the setting value with.
(Initial value: 0)
Pressto validate it.
3. Press twice to go back to the indicated value
display.
34
34
5 EXPLANATION OF INDICATED- VALUE-RELATED FUNCTIONS
5EXPLANATION OF INDICATED-
VALUE-RELATED FUNCTIONS
Ch
1
Ch 2
H
I
LO
H
I
LO
FNC
H
I
Ch
1
Ch 2
OK
LO
PEAK
HOLD
H
I
LO
HILO
ESC
5-1.A/D conversion speed select
The A/D conversion speed can be selected from 300 times/sec. or 3000 times/sec.
≪Setting value≫
0:300 times/sec.1:3000 times/sec.
If the A/D conversion speed is changed, the processing shown below will follow it.
● Comparison
According to the processing of load data, the comparison processing also follows the A/D
conversion speed.
Chapter
● Digital filter
The unit time in frequency setting changes.
Therefore, the average processing time will also change.
When 30 times is set at 3000 times/sec., 30 times ÷ 3000 times/sec.= 10msec.
When 30 times is set at 300 times/sec., 30 times ÷ 300 times/sec. = 100msec.
See "5-3.Digital filter" on page 37.
● I/O scan time
For example, the delay time of output signals changes.
The delay time is 0.7 msec at 3000 times/sec, and 7 msec (10 times) at 300 times/sec.
● BCD output interface (option)
The fastest setting of update rate is 300 times/sec when the A/D conversion speed is 300 times/
sec.
● D/A converter (option) output
Updating of load data follows the A/D conversion speed.
■A/D conversion speed select setting method
Select setting mode 3.
1.
Press → twice.
5
2. Set "A/D conversion speed."
Press five times. ⇒ "A/D conversion speed" is
displayed.
Press once, and present set value blinks.
Input the setting value with .
Press to validate it.
3. Press twice to go back to the indicated value display.
35
EXPLANATION OF INDICATED- VALUE-RELATED FUNCTIONS
35
Chapter
Ch
1
Ch 2
H
I
LO
H
I
LO
FNC
H
I
Ch
1
Ch 2
OK
LO
PEAK
HOLD
H
I
LO
H
I
LO
ESC
5 EXPLANATION OF INDICATED- VALUE-RELATED FUNCTIONS
5-2.Analog filter
This is a low-pass filter for filtering input signals from the strain gauge sensor, and canceling
unwanted noise components (a filter that passes only the signals with a specific frequency or less,
and attenuates signals with higher frequencies). With an increase in the cut-off frequency, the
response becomes faster, but noise components may be displayed. It can be set for each of the ch1
and ch2 inputs.
≪Setting value≫
0:3Hz
1:30Hz
2:300Hz
3:1kHz
■Analog filter setting method
Select setting mode 3.
1.
Press → twice.
5
EXPLANATION OF INDICATED- VALUE-RELATED FUNCTIONS
2. Set "Analog filter."
Press twice. ⇒ " Analog filter " is displayed.
Press , and present set value blinks.
Input the setting value with .
Press to validate it.(Initial value: 2)
3. Press twice to go back to the indicated value display.
36
36
5 EXPLANATION OF INDICATED- VALUE-RELATED FUNCTIONS
Ch
1
Ch 2
H
I
LO
H
I
LO
FNC
H
I
Ch
1
Ch 2
OK
LO
PEAK
HOLD
H
I
LO
HILO
Selected ch display
ESC
5-3.Digital filter
The digital filter is a function for reducing fluctuations of the indicated value by means of a moving
average of data converted from analog to digital. With an increase in the number of filterings, the
indicated value becomes more stable, but the response to inputs becomes slower. It can be set for
each of the ch1 and ch2 values.
≪Setting value≫
OFF(0)、2~999 [times]
* One time corresponds to one sampling for A/D conversion.
■Digital filter setting method
Select setting mode 3.
1.
Press → twice.
2. Set "Digital filter."
Press once. ⇒ "Digital filter" is displayed.
*It changes according to
to execute the
setting of ch2.
Press , and the highest digit blinks.
Input the setting value with .
Press to validate it.(Initial value: 000)
3. Press twice to go back to the indicated value display.
Chapter
5
37
EXPLANATION OF INDICATED- VALUE-RELATED FUNCTIONS
37
Chapter
H
I
Ch
1
Ch 2
OK
LO
PEAK
HOLD
H
I
LO
HILO
ZERO
H
I
Ch
1
Ch 2
OK
LO
PEAK
HOLD
H
I
LO
HILO
H
I
Ch
1
Ch 2
OK
LO
PEAK
HOLD
H
I
LO
H
I
LO
H
I
Ch
1
Ch 2
OK
LO
PEAK
HOLD
H
I
LO
H
I
LO
ESC
・When Digital Zero is performed, if the difference from the zero calibration
point exceeds the Digital Zero limit, the Digital Zero limit error ( )
will result.
Also, only the Digital Zero limit is subtracted.
・If the digital offset is set, even if Digital Zero is executed, zero will not result.
(Indicated value = -Setting value of digital offset)
・When Digital Zero reset is performed, the condition previous to correction by
Digital Zero is restored.
Point
5 EXPLANATION OF INDICATED- VALUE-RELATED FUNCTIONS
5-4.Digital Zero
This function is to zero the indicated value.
■Digital Zero / Digital Zero reset by means of keys
Digital Zero can be executed for the ch1 value and ch2 value individually or at one time.
Digital Zero reset can also be executed.
1. Press on the normal measurement screen.
2. The display becomes as shown on the right-hand
side, and execution is waited for on both ch1 and ch2.
Press once.
5
EXPLANATION OF INDICATED- VALUE-RELATED FUNCTIONS
3. The display becomes as shown on the right-hand
side, and execution is waited for only on ch1.
Pressonce.
4. The display becomes as shown on the right-hand
side, and execute-waits only for ch2.
Press once more to go back to the screen in 2.,
for execute-wait of both ch1 and ch2.
5. During execute-wait as in 2. - 4.:
Pressonce to zero the indicated value for execute-wait.
Pressonce to perform Digital Zero reset of the indicated value for execute-
wait.
Pressonce to cancel the state for execute-wait and go back to the indicated
value display state.
38
38
5 EXPLANATION OF INDICATED- VALUE-RELATED FUNCTIONS
D/Z
OFF
ON
Keep ON for 1.0 msec or more.
Ch
1
Ch 2
HILO
HILO
FNC
H
I
Ch
1
Ch 2
OK
LO
PEAK
HOLD
H
I
LO
HILO
ESC
■Digital Zero by means of the external signal "D/Z input"
The instant when D/Z and COM2 of the input/output connector are short-circuited from an open
state (ON edge), Digital Zero functions to zero both the ch1 value and ch2 value at one time.
For executing Digital Zero for the ch1 value and ch2 value individually, see "5-5.D/Z
synchronous mode" on page 39.
CAUTION
Digital-Zeroed points are saved in SRAM.
When the battery is executed, the Digital Zero is reset.
Chapter
5-5.D/Z synchronous mode
In execution of Digital Zero by means of the external input signal, Digital Zero can be executed for
the ch1 value and ch2 value individually or at one time.
≪Setting value≫
0:ch1 and ch2 synchronous When the external input "D/Z-1 (B11) is turned ON,
Digital Zero functions on both ch1 and ch2 at the
same timing.
1:ch1 and ch2 independentWhen the external input "D/Z-1 (B11) is turned ON,
Digital Zero functions on ch1, and when the external
input "D/Z-2 (B12) is turned ON, Digital Zero
functions on ch2.
■D/Z synchronous mode setting method
Select setting mode 3.
1.
Press→twice.
5
2. Set "D/Z synchronous mode."
Pressfour times ⇒ "D/Z synchronous mode" is
displayed.
Press, and present set value blinks.
Input the setting value with .
Pressto validate it.
39
3. Press twice to go back to the indicated value display.
39
EXPLANATION OF INDICATED- VALUE-RELATED FUNCTIONS
Chapter
Ch
1
Ch 2
H
I
LO
H
I
LO
FNC
0000
Range
-
00 ~ 99
Time
0.0 ~ 9.9
《
setting value》
.
H
I
Ch
1
Ch 2
OK
LO
PEAK
HOLD
HILO
H
I
LO
Selected ch display
【second】
【count】
ESC
5
5 EXPLANATION OF INDICATED- VALUE-RELATED FUNCTIONS
5-6.Zero Tracking
This function is to automatically track and correct gradual changes in the zero point due to drifts
(phenomenon in which outputs fluctuate slowly due to temperature change or change with time),
etc.
≪Setting value≫
Zero Tracking (time)0.0~9.9 [second]
Zero Tracking (range)00~99[count]
■Zero Tracking setting method
Select setting mode 3.
1.
Press→twice.
2. Set "Zero Tracking."
Press twice. ⇒ " Zero Tracking " is
displayed.
EXPLANATION OF INDICATED- VALUE-RELATED FUNCTIONS
*It changes according to
to execute the
setting of ch2.
Press, and the highest digit blinks.
①Set " Zero Tracking (time)."
Input the setting value with .
Move to the lower digits with .
(Initial value: 0.0)
②Set " Zero Tracking (rande)."
Input the setting value with .
Move to the lower digits with .(Initial value: 00)
Press to validate it.
3. Press twice to go back to the indicated value display.
40
40
5 EXPLANATION OF INDICATED- VALUE-RELATED FUNCTIONS
・When displacement of the zero point is within the set count of tracking and it
continues more than the set time, it is automatically made zero by Zero
Tracking function.
・The time (tracking delay) is set in the range of 0.1 - 9.9 sec., and the band
(tracking band) is set in the range of 01 to 99.
If the time is set at 0.0 sec. and the band at 00, the zero tracking function does
not work.
Boundary of zero track
DELAY
Indicated value
+COUNT
-COUNT
BAND
+
0
-
DELAY
Band=count×2
From the point when it returned within the range,
counting will be resumed.
When the time is 0.0 sec. and the width is 00 markings, stable is not detected.
Stability detection by Motion Detect is closely related to the SI/F print (stable
value) function and calculation comparison timing. For details, see "6-
5.Calculation comparison timing" on page 48 and "■ SI/F print" on page 62.
Point
Chapter
CAUTION
● Zero Tracking is equal to Digital Zero functioning automatically. Therefore, the
Digital Zero limit is also monitored.
● Zero Tracking does not function if the indicated value has already exceeded the
tracking band.
5-7.Motion Detect
Set the parameters to detect stability of calculated values.
If the difference between the current indicated value and the 50-msec-previously indicated value
becomes less than the set count and that condition continues for more than the set time, the
5 EXPLANATION OF INDICATED- VALUE-RELATED FUNCTIONS
Setting example
Motion Detect (time)0.4 sec.
Motion Detect (range)10
5
EXPLANATION OF INDICATED- VALUE-RELATED FUNCTIONS
When the condition of dn ≦ set count (10) continues for the set time (0.4 sec.) or more
(d1 ≦ 10, and d2 ≦ 10, and ..., and d8 ≦ 10), the indicated value is regarded as stable.
■Motion Detect setting method
Select setting mode2.
1.
Press→once.
2. Set "Motion Detect."
Press once. ⇒ "Motion Detect" is displayed.
Press , and the highest digit blinks.
① Set " Motion Detect time."
Input the setting value with.
Move to the lower digits with.
(Initial value: 0.0)
② Set " Motion Detect range."
Input the setting value with .
Move to the lower digits with .
42
42
(Initial value: 00)
Press to validate it.
3. Press twice to go back to the indicated value display.
5 EXPLANATION OF INDICATED- VALUE-RELATED FUNCTIONS
Ch
1
Ch 2
H
I
LO
H
I
LO
FNC
H
I
Ch
1
Ch 2
OK
LO
PEAK
HOLD
H
I
LO
H
I
LO
ESC
5-8.Display frequency
Set the per-second display update frequency of the indicated value. However, the A/D conversion
speed does not change.
≪Setting value≫
0: 3 times
1: 6 times
2: 13 times
3: 25 times
■Display frequency setting method
Select setting mode 2.
1.
Press→once.
2. Set "Display frequency."
Pressthree times ⇒ "Display frequency" is
displayed.
Press, and present set value blinks.
Input the setting value with.
Pressto validate it.(Initial value: 3)
3. Presstwice to go back to the indicated value display.
Set the calculation method for the ch1 value and ch2 value.
0:ch1+ch2The ch2 value is added to the ch1 value.
1:ch1-ch2The ch2 value is subtracted from the ch1 value.
2:|ch1-ch2|The absolute value of the difference between the ch1 value and
ch2 value is calculated.
3:High selectA larger value is always selected.
(When the ch1 value ≧ ch2 value, the ch 1 value results, and when
the ch1 value < ch2 value, the ch2 value results.)
4:Low selectA smaller value is always selected.
(When the ch1 value ≦ ch2 value, the ch 1 value results, and when
the ch1 value > ch2 value, the ch2 value results.)
6
■Calculation mode setting method
EXPLANATION OF COMPARISON FUNCTIONS
Select setting mode1.
1.
Press once.
2. Set "Calculation mode."
Press four times ⇒ "Calculation mode" is
displayed.
Press , and the highest digit blinks.
Input the setting value with .
(Initial value:0)
Press to validate it.
3. Press twice to go back to the indicated
value display.
44
44
6 EXPLANATION OF COMPARISON FUNCTIONS
With ±LOAD, OFL1, or OFL2 on ch1 or ch2, HI/LO of the error channel is forcedly
turned on.
Also, HI/LO of the calculated value is forcedly turned on, and OK is forcedly
turned off.
Point
6-2.HI/LO limit comparison of calculated value /
ch1 value / ch2 value
Set the HI/LO limit of calculated value/ch1 value/ch2 value.
Calculated value-99999 ~ 99999
Ch1 value / ch2 value -19999 ~ 19999
<HI/LO output conditions>
・HIIndicated value > HI limit
・LOIndicated value < LO limit
<OK output conditions (calculated value)>
・OKLO limit ≦ Indicated value ≦ HI limit
* The OK output conditions basically depend on the calculated value, but it can also be
output as the ch1 value and ch2 value are intertwined. See "6-6. Comparison OK
output conditions" on page 49.
■HI/LO limit setting method
See "≪HI/LO limit setting mode function≫" on page 21.
6-3.Hysteresis
This function is to allow a margin for timing at which HI/LO of the HI/LO limit comparison is
turned off. Normally, HI is turned on when the indicated value exceeds the HI limit and is turned off
when the indicated value falls below it. However, by setting the hysteresis, HI is turned off when
the indicated value falls below the HI limit further lowered by the hysteresis value. This function is
effective to prevent chattering (phenomenon in which signals repeat turning on/off immediately
after switching of relay contacts and switch contacts) in such a case where signals fluctuate
(vibrate) subtly.
The setting is common to the calculated value, ch1 value, and ch2 value, which operates
individually.
Chapter
6
<Hysteresis range>
0~9999
<HI/LO output conditions>
・HI
ON conditionsIndicated value > HI limit
45
OFF conditions Indicated value ≦ (HI limit setting value-Hysteresis setting value)
EXPLANATION OF COMPARISON FUNCTIONS
45
6 EXPLANATION OF COMPARISON FUNCTIONS
High limit value
Low limit value
+
0
Hysteresis range
Time
HI output
HI status display
LO output
LO status display
OK status display
Output ON
Status ON
Output OFF
Status OFF
Status ON Status ON
Output ON
Status ON
Output ON
Status ON
-
Output OFF
Status OFF
Output OFF
Status OFF
Indicated value
OK output
(Example:HI output, LO output, OK output)
Ch
1
Ch 2
H
I
LO
H
I
LO
FNC
H
I
Ch
1
Ch 2
OK
LO
PEAK
HOLD
H
I
LO
H
I
LO
ESC
The setting value of hysteresis is common to the HI limit and LO limit of the
calculated value, ch1 value, and ch2 value.
Point
・LO
ON conditions Indicated value < LO limit
OFF conditions Indicated value ≧
・Hysteresis operation
(LO limit setting value-Hysteresis setting value)
Chapter
6
EXPLANATION OF COMPARISON FUNCTIONS
■Hysteresis setting method
Select setting mode1.
1.
Press once.
2. Set "Hysteresis."
Pressthree times ⇒ "Hysteresis" is displayed.
Press, and the highest digit blinks.
Input the setting value with.
Move to the lower digits with.
(Initial value:0000)
Pressto validate it.
46
46
3. Presstwice to go back to the Indicated value display.
6-4.Near Zero
Near Zero ON/OFF is closely related to the SI/F print (stable value) function and
calculation comparison timing.
For details, see "6-5. Calculation comparison timing" on page 48, and "■ SI/F
print" on page 62.
Point
Ch
1
Ch 2
H
I
LO
H
I
LO
FNC
H
I
Ch
1
Ch 2
OK
LO
PEAK
HOLD
H
I
LO
H
I
LO
ESC
Near Zero operates by absolute value.
Point
This function is to detect that the indicated value (calculated value) is Near Zero.
Near Zero ON:|Calculated value| ≦ Setting value of Near Zero
Near Zero OFFF:|Calculated value| > Setting value of Near Zero
≪Setting value≫
0~99999
6EXPLANATION OF COMPARISON FUNCTIONS
■Near Zero setting method
Select setting mode1.
1.
Pressonce.
2. Set "Near Zero."
Presstwice ⇒ "Near Zero" is displayed.
Press, and the highest digit blinks.
Input the setting value with.
Move to the lower digits with.(Initial value:100)
Pressto validate it.
3. Presstwice to go back to the indicated value display.
Chapter
6
47
EXPLANATION OF COMPARISON FUNCTIONS
47
6 EXPLANATION OF COMPARISON FUNCTIONS
Ch
1
Ch 2
H
I
LO
H
I
LO
FNC
H
I
Ch
1
Ch 2
OK
LO
PEAK
HOLD
H
I
LO
H
I
LO
Undefined
Calculation comparison OK
0 0 0 0
Calculation comparison timing
Undefined
0:All time
1:Stable
2:Near Zero OFF
3:Stable and Near Zero OFF
4: Hold
output condition
ESC
6-5.Calculation comparison timing
Set the operating conditions of calculation HI/LO limit comparison. Select the conditions from the
following.
0:All timeHI/LO limit comparison is always performed.
1:StableHI/LO limit comparison is performed when the calculated
value is stable.
2:Near Zero OFFHI/LO limit comparison is performed when Near Zero is
off.
3:Stable and Near Zero OFF HI/LO limit comparison is performed when the calculated
value is stable and Near Zero is off.
4:HoldHI/LO limit comparison is performed when a hold is
fixed.
■Calculation comparison timing setting method
Chapter
6
EXPLANATION OF COMPARISON FUNCTIONS
1.
Select setting mode1.
Pressonce.
2. Set "Calculation comparison timing."
Pressonce. ⇒ "Calculation function
select" is displayed.
Press, and the highest digit blinks.
Input the setting value with.
(Initial value:0)
Pressto validate it.
3. Presstwice to go back to the indicated
value display.
48
48
6 EXPLANATION OF COMPARISON FUNCTIONS
Ch
1
Ch 2
H
I
LO
H
I
LO
FNC
H
I
Ch
1
Ch 2
OK
LO
PEAK
HOLD
H
I
LO
H
I
LO
H
I
Ch
1
Ch 2
OK
LO
PEAK
HOLD
HILO
H
I
LO
Undefined
Calculation comparison timing
0 0 0 0
Calculation comparison OK
Undefined
0:Calculated value only
1:Calculated value & ch1
2:Calculated value & ch2
3:Calculated value & ch1 & ch2
output conditions
ESC
6-6.Comparison OK output conditions
Conditions to turn on the external output "Calculation OK output" can be set.
0:Calculated value onlyThe output turns on when calculation HI limit/LO
limit is off.
1: Calculated value & ch1The output turns on when calculation HI limit/LO
limit and ch1 HI limit/LO limit are all off.
2:Calculated value & ch2The output turns on when calculation HI limit/LO
limit and ch2 HI limit/LO limit are all off.
3:Calculated value & ch1 & ch2The output turns on when calculation HI limit/LO
limit, ch1 HI limit/LO limit, and ch2 HI limit/LO limit
are all off.
■Comparison OK output conditions setting method
Select setting mode1.
1.
Pressonce.
2. Set "Comparison OK output conditions setting ."
Pressonce. ⇒ "Calculation function select"
is displayed.
Press, and the highest digit blinks.
Pressonce, and the lower digits blinks.
Input the setting value with.
(Initial value:0)
Pressto validate it.
Chapter
6
3. Presstwice to go back to the indicated
value display.
49
EXPLANATION OF COMPARISON FUNCTIONS
49
7 EXPLANATION OF HOLD FUNCTIONS
7EXPLANATION OF
HOLD FUNCTIONS
7-1.Hold functions of calculated values
The hold functions are to take out a specific point of calculated values for HI/LO limit comparison.
Here, the operation of each hold will be described in detail.
Hold modeHold section
0:OFF
1:Sample0:All section 1:External signal
2:Peak
4:Average
■Flow of hold measurement
Chapter
Select a hold you want to use.
1.
7
2. If the hold has the setting of section, set a section you want to use.
EXPLANATION OF HOLD FUNCTIONS
For details, see "7-3.Setting of hold section" on page 54.
3. Set the special setting value required for the hold used.
4. After confirmation of wiring, perform hold control by external input, and start
measurement.
None
0:All section 1:External signal 2:External signal+Time3:Bottom
50
50
7-2.Setting of hold mode
Ch
1
Ch 2
H
I
LO
H
I
LO
FNC
H
I
Ch
1
Ch 2
OK
LO
PEAK
HOLD
HILO
H
I
LO
Hold section
Hold mode
0 0 0 1
Calculation mode
Auto reset
《Setting value》
0:OFF
1: Sample
2: Peak
3: Bottom
4: Average
ESC
Indicated
t2
t3
t1
Internal caleulated value
Hold Section
Fixed
t
+
T/H
HI-LO Limit
Judging Output
H/E
OFF
ON
OFF
ON
OFF
ON
HOLD
OFF
ON
Button
HOLD
Statu s Displa y
PEAK
Stat us Dis play
ONBlinkBlink
OFF
Val ue
■Hold mode setting method
1. Select setting mode 1.
Pressonce.
2. Set "Hold mode."
Press four times ⇒ "Calculation hold function
select " is displayed.
Press , and the highest digit blinks.
Move to the lower digits with .
Input the setting value with.
(Initial value:0001)
Validate it with .
7 EXPLANATION OF HOLD FUNCTIONS
51
3. Press twice to go back to the indicated value
■Sample hold
The start point of the specified section is held.
The section is specified by the setting of "all section" or "external signal."
(Example) All section sample hold
display.
Chapter
7
t1: A delay time between the instant
when the T/H signal is inputted and the
instant when the indicated value is held
Approximately 0.7ms (MAX.)
t2: A delay time between the instant
when the T/H signal is released and the
instant
when the indicated value returns to
tracking
Approximately 0.7ms (MAX.)
t3: A minimum reset signal width
required for releasing the hold
Approximately 0.7ms (MIN.)
The delay time and signal width
expressed apply when the A/D
conversion speed is 3000 times/sec.
They become 10-times when the A/D
conversion speed is 300 times/sec.
51
EXPLANATION OF HOLD FUNCTIONS
7 EXPLANATION OF HOLD FUNCTIONS
Indicated value
t3
Internal calculated
Detection
t
+
T/H
HI-LO Limit
Judging Output
H/E
Hold Section
t2
t1
Fixed
SECTION
OFF
ON
OFF
ON
OFF
ON
OFF
ON
OFF
ON
ON
BlinkON
OFFON
OFF
OFF
OFF
HOLD
Button
HOLD
Status Display
PEAK
Status Display
value
Indicated
t2
t3
Internal calculated
Detection・Hold Section
Undetermind
t
+
T/H
HI-LO Limit
Judging Output
H/E
Detection・Hold Section
t1t2t1
t3
OFF
ON
OFF
ON
OFF
ON
OFF
ON
HOLD
Status Display
PEAK
Status Display
ON
BlinkBlink
OFFON
OFF
OFF
OFF
OFF
OFF
HOLD
Button
Section (Note)
Undetermind
Section (Note)
Val ue
value
■Peak hold
The maximum value in the positive direction (peak value) of the specified section is held.
The section is specified by the setting of "all section," "external signal," or "external signal + time."
(Example) Externally specified section peak hold
t1: A delay time between the instant
when the SECTION signal is inputted
and the instant when the value to be
held is detected
Approximately 0.7ms (MAX.)
t2: A delay time between the instant
when the SECTION signal is released
and the instant when the value to be
held is determined
Approximately 0.7ms (MAX.)
t3: A minimum reset signal width
required for releasing the hold
Approximately 0.7ms (MIN.)
The delay time and signal width
expressed apply when the A/D
conversion speed is 3000 times/sec.
They become 10-times when the A/D
conversion speed is 300 times/sec.
■Bottom hold
Chapter
7
EXPLANATION OF HOLD FUNCTIONS
52
52
The maximum value in the negative direction (bottom value) of the specified section is held.
The section is specified by the setting of "all section," "external signal," or "external signal + time."
(Example) All section Bottom hold
Note: During the undetermined section, the judging output varies with fluctuations in the input waveform.
However, the H/E output remains on during the undetermined section. Read the judging result when the
indicated value becomes stable (immediately before the T/H signal rises).
t1: A delay time between the instant
when the T/H signal is inputted and
the instant when the indicated value
is held
Approximately 0.7ms (MAX.)
t2: A delay time between the instant
when the T/H signal is released and
the instant
when the indicated value returns to
tracking
Approximately 0.7ms (MAX.)
t3: A minimum reset signal width
required for releasing the hold
Approximately 0.7ms (MIN.)
The delay time and signal width
expressed apply when the A/D
conversion speed is 3000 times/sec.
They become 10-times when the A/D
conversion speed is 300 times/sec.
7 EXPLANATION OF HOLD FUNCTIONS
Internal calculated value
Hold Value
t
+
T/H
Limit Judging
H/E
Fixed
SECTION
OFF
ON
OFF
ON
OFF
ON
OFF
ON
Undetermind
t1t2t3
DetectionHold Sect ion
OFF
ON
HOLD
Button
HI-LO
HOLD
Status Display
PEAK
Status Display
ON
BlinkON
OFFONOFF
OFF
OFF
Section (Note)
Output
■Average hold
The average of the sampling values over the specified section is calculated and updated, and then
held.
The section is specified by the setting of "all section," "external signal," or "external signal + time."
(Example) Externally specified section average value hold
t1: A delay time between the instant
when the SECTION signal is inputted
and the instant when the value to be held
is detected
Approximately 0.7ms (MAX.)
t2: A delay time between the instant
when the SECTION signal is released
and the instant when the value to be held
is determined
Approximately 0.7ms (MAX.)
t3: A minimum reset signal width
required for releasing the hold
Approximately 0.7ms (MIN.)
The delay time and signal width
expressed apply when the A/D
conversion speed is 3000 times/sec.
They become 10-times when the A/D
conversion speed is 300 times/sec.
The detection section is limited by the maximum average value detection time set according to the
Chapter
"average sample number." If the maximum average value detection time is exceeded, detection is
automatically terminated, and the average value of that time is held.
7
■Average sample number
In the average hold, the maximum average value detection time for average processing by each
sampling depends on the A/D conversion speed as follows:
When the A/D conversion speed is 300 times/sec:50sec.
When the A/D conversion speed is 3000 times/sec: 5sec.
Then, if this average sample number is set at 2 or more, the sampling data used for average
calculation becomes the representative value of the sampling values of the set number (average
obtained according to the number). By this setting, the average value detection time can be
elongated, but error causes by average calculation increase.
Example) When the A/D conversion speed is 3000 times/sec,
Maximum average value detection time = Average sample number × 10 sec.,
and also the number of average updates will decrease.
Number of average updates = 3000 times/sec. ÷ Average sample number
≪Setting value≫
53
1~999 [times]
EXPLANATION OF HOLD FUNCTIONS
53
7 EXPLANATION OF HOLD FUNCTIONS
Ch
1
Ch 2
H
I
LO
H
I
LO
FNC
H
I
Ch
1
Ch 2
OK
LO
PEAK
HOLD
H
I
LO
H
I
LO
ESC
Ch
1
Ch 2
H
I
LO
H
I
LO
FNC
H
I
Ch
1
Ch 2
OK
LO
PEAK
HOLD
HILO
H
I
LO
Hold section
Hold mode
0 0 0 1
Calculation mode
Auto reset
0: All section
1: External signal
2: External signal + time
* This setting is required only
than "0: OFF."
when the hold mode is other
ESC
◇Average sample number setting method
1. Select setting mode 1.
Pressonce.
2. Set "Average sample number."
Press six times ⇒ "Average sample number "
is displayed.
Press , and the highest digit blinks.
Input the setting value with. (Initial value:1)
Move to the lower digits with .
Press to validate it.
3. Press twice to go back to the indicated value display.
7-3.Setting of hold section
■Hold section setting method
Chapter
7
EXPLANATION OF HOLD FUNCTIONS
Select setting mode 1.
1.
Press once.
2. Set "Hold section."
Press four times ⇒ "Calculation hold function
select " is displayed.
Press , and the highest digit blinks.
Press twice to make the digit of the Hold
section blink. (Initial value: 0001)
Input the setting value with .
Press to validate it.
3. Press twice to go back to the indicated value
display.
54
54
7 EXPLANATION OF HOLD FUNCTIONS
Indicated value
t2
t3
Internal caluculated value
Detection
・
Hold Section
Undetermind
t
+
T/H
HI-LO Limit
Judging Output
H/E
Detection・Hold Section
t1
t2t1
t3
Undetermind
OFF
ON
OFF
ON
OFF
ON
OFF
ON
HOLD
Status Display
PEAK
Status Display
ON
BlinkBlink
OFFON
OFF
OFF
OFF
OFF
OFF
HOLD
Button
Section(Note)
Section(Note)
■All section
By this method, the hold detection section is specified by the T/H signal.
Detection starts with the T/H signal ON to perform each hold operation.
Control is performed simply by the T/H signal without using the SECTION signal.
Therefore, there are tracking sections and detection sections, but there are no hold deciding
sections.
(Example) All section peak hold
t1: A delay time between the instant
when the T/H signal is inputted and
the instant when the indicated value
is held
Approximately 0.7ms (MAX.)
t2: A delay time between the instant
when the T/H signal is released and
the instant
when the indicated value returns to
tracking
Approximately 0.7ms (MAX.)
t3: A minimum reset signal width
required for releasing the hold
Approximately 0.7ms (MIN.)
The delay time and signal width
expressed apply when the A/D
conversion speed is 3000 times/sec.
They become 10-times when the A/
D conversion speed is 300 times/sec.
Note:During the undetermined sections, the judging output varies with fluctuations in the input waveform.
However, the H/E output remains on during the undetermined sections. Read the judging result when
the indicated value becomes stable (immediately before the T/H signal rises).
Chapter
7
55
EXPLANATION OF HOLD FUNCTIONS
55
7 EXPLANATION OF HOLD FUNCTIONS
Indicated value
t3
Internal calculated
Detection
t
+
T/H
HI-LO Limit
Judging Output
H/E
Hold Section
t2
t1
Fixed
SECTION
OFF
ON
OFF
ON
OFF
ON
OFF
ON
OFF
ON
ON
BlinkON
OFFON
OFF
OFF
OFF
HOLD
Button
HOLD
Status Display
PEAK
Status Display
value
■External signal
By this method, the hold detection section is specified by the SECTION signal.
Detection starts with the SECTION signal ON, and ends with the SECTION signal OFF to perform
each hold operation.
The hold value is maintained until the hold reset signal is turned on.
The hold is released by turning on the T/H signal as a reset signal.
There are tracking sections, detection sections, and hold deciding sections.
tant when the T/H signal is turned on.
(Example) Externally specified section peak hold
t1: A delay time between the instant
when the SECTION signal is inputted
and the instant when the value to be
held is detected
Approximately 0.7ms (MAX.)
t2: A delay time between the instant
when the SECTION signal is released
and the instant when the value to be
held is determined
Approximately 0.7ms (MAX.)
Chapter
7
EXPLANATION OF HOLD FUNCTIONS
again without turning on the T/H signal while the hold is fixed, the hold is released and re-
detection is started. For details, see "7-4.Auto reset function" on page 58.
t3: A minimum reset signal width
required for releasing the hold
Approximately 0.7ms (MIN.)
The delay time and signal width
expressed apply when the A/D
conversion speed is 3000 times/sec.
They become 10-times when the A/D
conversion speed is 300 times/sec.
By turning on the SECTION signal
56
56
7 EXPLANATION OF HOLD FUNCTIONS
Indicated value
t3
Internal calculatde value
Detection
t
+
HI-LO Limit
Judging Output
H/E
Hold Section
t2
t1
Fixed
SECTION
T/H
OFF
ON
OFF
ON
OFF
ON
OFF
ON
OFF
ON
OFF
ON
ON
BlinkON
OFFON
OFF
OFF
OFF
HOLD
Button
Detection time
HOLD
Status Display
PEAK
Status Display
Ch
1
Ch 2
H
I
LO
H
I
LO
FNC
H
I
Ch
1
Ch 2
OK
LO
PEAK
HOLD
HILO
H
I
LO
ESC
■External signal + time
By this method, the hold detection section is specified as a timer. Detection starts with the
SECTION signal ON, and ends when the set time (detection time) has elapsed.
The hold value is maintained until the hold reset signal is turned on.
The hold is released by turning on the T/H signal as a reset signal.
There are tracking sections, detection sections, and hold deciding sections.
(Example) External signal+ Time specified section peak hold
t1: A minimum SECTION signal width
required for beginning detection
Approximately 0.7ms (MIN.)
t2: A delay time between the instant
when the detection time is ended and
the instant when the value to be held is
determined
Approximately 0.7ms (MAX.)
t3: A minimum reset signal width
required for releasing the hold
Approximately 0.7ms (MIN.)
57
By turning on the SECTION signal again without turning on the T/H signal while the hold is
fixed, the hold is released and re-detection is started.
For details, see "7-4.Auto reset function" on page 58.
■Detection time
Set the hold detection time.
《Setting value》
0.01~9.99 [sec.]
◇ Detection time setting method
1. Select setting mode 1.
Pressonce.
2. Set "Detection time."
Pressfive times ⇒ "Detection time" is displayed.
Press, and the highest digit blinks.
Move to the lower digits with.(Initial value:1)
Input the setting value with.
Press to validate it.
3. Press twice to go back to the indicated value display.
The delay time and signal width
expressed apply when the A/D
conversion speed is 3000 times/sec.
They become 10-times when the A/D
conversion speed is 300 times/sec.
Chapter
7
EXPLANATION OF HOLD FUNCTIONS
57
7 EXPLANATION OF HOLD FUNCTIONS
SECTION
T/H
H/E
Example) Auto reset in externally specified section peak hold
Upon turning-ON of the SECTION signal,
hold reset is performed and detection starts.
Control can be performed without inputting the T/H
signal.
However, a hold value is maintained until the
SECTION signal turns ON.
When auto reset is on in externally
SECTION
H/E
T/H
When auto reset is off in externally
SECTION
H/E
T/H
SECTION
H/E
T/H
SECTION
H/E
T/H
When auto reset is on in externally
When auto reset is off in externally
If the SECTION signal
chatters, the point shifts
from the proper one.
The first hold value is
maintained until T/H is
inputted.
If the SECTION signal
chatters, the point shifts from
the proper one.
The first hold value is
maintained until T/H is
inputted.
specified section sample hold
specified section peak hold
specified section peak hold
specified section sample hold
7-4.Auto reset function
It is factory-set that each hold is automatically reset at the start of the detection section, and
therefore, simple control can be performed by one signal without releasing the hold.
However, when the auto reset functions, if the signal causes chattering, the point that should be
taken originally may deviate. Switching is provided so as to prevent the next detection from being
performed until the T/H signal is turned on even if the SECTION signal is turned on again.
《Setting value》
0:OFFAfter a hold is fixed, the next detection is not started until the T/H
signal is turned on even if the SECTION signal is turned on.
Chapter
7
1:ONAfter a hold is fixed, the next detection is started when the SECTION
signal is turned on again even if the T/H signal is not turned on. This is
the initial setting (factory default setting).
EXPLANATION OF HOLD FUNCTIONS
58
58
■Auto reset setting method
Ch
1
Ch 2
H
I
LO
HILO
FNC
H
I
Ch
1
Ch 2
OK
LO
PEAK
HOLD
H
I
LO
H
I
LO
Hold section
Hold mode
0 0 0 1
Calculation mode
Auto reset
《Setting value》
0:OFF
1:ON
* This setting is required when the
hold section is "external signal"
or "external signal + time."
ESC
Select setting mode 1.
1.
Pressonce.
2. Set "Auto reset."
Pressfour times ⇒ "Calculation hold
function select " is displayed.
Press, and the highest digit blinks.
Pressthree times to make the digit of
the auto reset blink. (Initial value: 0001)
Input the setting value with.
Press to validate it.
3. Presstwice to go back to the
indicated value display.
7 EXPLANATION OF HOLD FUNCTIONS
Chapter
7
EXPLANATION OF HOLD FUNCTIONS
59
59
8 STANDARD INTERFACE
8STANDARD INTERFACE
8-1.External I/O
■Connector pin assignment
A1COM1B1COM1
A2OUTHI(Calculated)B2OUTALM(ch2)
A3OUTOK(Calculated)B3OUTH/E
A4OUTLO(Calculated)B4OUT
A5OUTHI(ch1)B5OUT
A6OUTLO(ch1)B6OUTRUN
A7COM1B7COM2
Chapter
8
STANDARD INTERFACE
A8OUTALM(ch1)B8INSECTION
A9OUTHI(ch2)B9INT/H
A10OUTLO(ch2)B10IN
A11OUTSI/FB11IND/Z-1
A12OUTSI/FB12IND/Z-2
Compatible connector: FCN-361J024-AU(manufactured by FUJITSU COMPONENT or equivalent)
Connector cover:FCN-360C024-B(manufactured by FUJITSU COMPONENT or equivalent)
(connector & cover optional type: CN50)
■About outputs
A1, A7, B1Common for external output signals.
A2Outputs the calculated value HI limit signal.
A3Outputs the calculated value OK signal.
A4Outputs the calculated value LO limit signal.
A5Outputs the ch1 value HI limit signal.
A6Outputs the ch1 value LO limit signal.
→
" ■ External I/O connection " on page 14
"6. EXPLANATION OF COMPARISON
→
FUNCTIONS " on page 44
A8Outputs the ch1 value alarm signal.
A9Outputs the ch2 value HI limit signal.
A10Outputs the ch2 value LO limit signal.
B2Outputs the ch2 alarm signal.
B3Outputs the hold complete signal.
B6Outputs the normal run signal.
→
"4. CALIBRATION PROCEDURE " on page 23
→ "6. EXPLANATION OF COMPARISON
FUNCTIONS " on page 44
→
"4. CALIBRATION PROCEDURE " on page 23
"7. EXPLANATION OF HOLD FUNCTIONS "
→
on page 50
→
" ■ About the RUN signal " on page 61
60
60
8 STANDARD INTERFACE
Refer to "11-7. Error/message list " on page 97 about the
contents of errors.
OFF
RUN
ON
OFF
RUN
ON
0.5sec. pulse
3.0sec.
Error factor removal
This signal turns off when an error occurs.
duty50%
* By the error factor removal, ON-OFF operation restarts in 3.0 sec.
* RUN turns off when the power is turned off.
A11, A12
2-wire serial interface for connecting a
UNIPULSE-manufactured printer, external
display, etc.
" ■ SI/F interface connection " on page 15
→
"8-2 SI/F(Serial interface) " on page 62
■About inputs
B7Common for external input signals.
B8, B9Inputs the hold control signal.
B11, B12Inputs the Digital Zero signal.
→
" ■ External I/O connection " on page 14
"7. EXPLANATION OF HOLD
→
FUNCTIONS " on page 50
→
"5-4. Digital Zero " on page 38
■About the RUN signal
The RUN output stops when an abnormal state of the indicated value is detected. The conditions
under which the output stops as in an abnormal state are as follows.
・±LOAD
・OFL1, OFL2
・Calibration error
・Digital Zero limit error
Chapter
8
STANDARD INTERFACE
61
61
8 STANDARD INTERFACE
Ch
1
Ch 2
H
I
LO
H
I
LO
FNC
H
I
Ch
1
Ch 2
OK
LO
PEAK
HOLD
H
I
LO
H
I
LO
000
Undefined
SI/F print
SI/F Indicated value
0: Calculated value
1: Calculated value
2:ch1/ch2
《Setting value》
/ch1
/ch2
format
ESC
8-2.SI/F(Serial interface)
■SI/F Indicated value format
For the SI/F, indicated value data exist in two areas: GROSS area and NET area.
Select the data.
《Setting value》
0:Calculated value/ch1 valueGROSS area:calculated valueNET area:ch1 value
1:Calculated value/ch2 valueGROSS area: calculated valueNET area: ch2 value
2:ch1 value/ch2 valueGROSS area: ch1 valueNET area: ch2 value
◇SI/F Indicated value format setting method
1. Select setting mode 2.
Press → once.
Chapter
8
STANDARD INTERFACE
62
■SI/F print
62
2. Set "SI/F Indicated value format."
Press twice ⇒ "SI/F function select " is
displayed.
Press , and the highest digit blinks.
Press once to make the digit of the SI/F
function select blink.
Input the setting value with.
(Initial value:0)
Press to validate it.
3. Presstwice to go back to the indicated value
display.
《Setting value》
0: NoneNo print command is sent.
1:Stable valueWhen stability of the calculated value is detected, a print command
2:Hold valueWhen hold is reset, a print command is sent to make the connected
Also, not-held calculated values are taken in the NET area.
is sent to the SI/F to make the connected printer print automatically.
When Near Zero is OFF, the calculated value is held until Near Zero
ON after stable ON.
Hold is reset in 1.5 sec. after Near Zero ON.
printer print the hold value automatically. In this case, the hold value
is forcedly taken in the GROSS area in sending.
◇SI/F print setting method
When stable value print is selected, be aware that automatic printing is not
performed in the following case.
・When Motion Detect is set in time: 0.0 sec, count: 00.
Also, the indicated value is not held in the following case.
・Hold mode:When any item other than none is selected.
Point
Ch
1
Ch 2
HILO
H
I
LO
FNC
SI/F print
0: None
H
I
Ch
1
Ch 2
OK
LO
PEAK
HOLD
H
I
LO
H
I
LO
000
Undefined
1:Stable value
SI00/F Indicated value
2:Hold value
《Setting value》
format
ESC
Indicated value
Near Zero
0
Stable
Near Zero
ON
ON
ONON
OFF
OFF
OFF
Automatic printing
↑
Hold
1.5seconds
Time
ON
*Hold function …
Indicated valueSI/F
Comparison output signal
+
Internal calculated value
Near Zero
0
Time
Stab le
Near Zero
ON
ON
ONON
OFF
OFF
OFF
Automatic printing
↑
Hold
1.5seconds
ON
-
Indicated
Internal calculated value
value
1. Select setting mode 2.
Press→once.
2. Set "SI/F print."
Press twice ⇒ "SI/F function select" is
displayed.
Press , and the highest digit blinks. Select the
digit of the SI/F print.
8 STANDARD INTERFACE
Input the setting value with.
(Initial value:0)
Press to validate it.
3. Presstwice to go back to the indicated value
display.
Chapter
・Operation of the indicated value hold function
8
STANDARD INTERFACE
63
For the "Stable" parameter, see "5-7. Motion Detect " on page 41, and for the "Near Zero"
parameter, see "6-4. Near Zero " on page 47.
63
9 OPTION INTERFACE
9OPTION INTERFACE
BCD data output
connector
● Internal transistor status
Output dataNegative Positive
0OFFON
1ONOFF
Vceo=30V(max)
● Output pin level
Output dataNegative Positive
0HL
1LH
Through logic switching (pin B14)
Vcc+5V
COM
Ic =30mA (max)
Inside
outside
Vext
F350
9-1.BCD interface
The BCD data output is an interface to extract the indicated value of the F350 as BCD data. This
interface is convenient to process controls, totals, records, etc., by connecting the F350 to a
computer, process controller, sequencer or the like.
The I/O and internal circuits are electrically insulated by photocoupler.
Chapter
9
OPTION INTERFACE
■Connection
●Output
The signal output circuit employs the TTL open collector output.
64
64
●Input
Transistor
TTL open collector output
Open OFF
ShortON
Vcc
+5V
IC=Approx.6mA
COM
Switch
Inside
Outside
F350
(ON when IN is HI)
CAUTION
● Do not apply external voltage to the signal input circuit.
The external element is required to withstand Ic=10mA.
●
Leakage from the external element is required to be 30μA or below
●
9 OPTION INTERFACE
.
■Connector pin assignment
A1
A2
A3
A4
A5
A6
A7
A8
A9
A10
A11
A12
A13
A14
A15
A16
*
OUT
OUT
OUT
OUT
OUT
OUT
OUT
OUT
OUT
OUT
OUT
OUT
IN
IN
-
COM
1B2
2B3
4B4
8B5
10B6
20B7
40B8
80B9
100B10
200B11
400B12
800B13
Data HoldB14
Data Switching 0B15
B1
B16
*
OUT
OUT
OUT
OUT
OUT
OUT
OUT
OUT
OUT
OUT
OUT
OUT
IN
IN
-
COM
1000
2000
4000
8000
10000
20000
40000
80000
Minus (Polarity)
OVER
STAB
STROBE
Logic Switching
Data Switching 1
Chapter
9
OPTION INTERFACE
65
Compatible connector: FCN-361J032-AU (manufactured by FUJITSU COMPONENT or equivalent)
Connector cover: FCN-360C032-B (manufactured by FUJITSU COMPONENT or equivalent)
(connector & cover optional type: CN51)
65
9 OPTION INTERFACE
Ch 2
HILO
Ch
1
H
I
LO
FNC
H
I
Ch
1
Ch 2
OK
LO
PEAK
HOLD
HILO
HILO
ESC
■Output data select
《Setting value》
0:Calculated valueThe calculated value is output.
1:ch1 valueThe ch1 value is output.
2:ch2 valueThe ch2 value is output.
3:External selectData selected by the data select input (A15, B15) is output.
Data select 1Data select 0Output data
OFFOFF
OFFONch1 value
ONOFFch2 value
ONON
* In the case of external selection, read after at least 2 cycles.
◇ Output data select setting method
1. Select setting mode 5.
Press→four times.
2. Set "Output data select."
Press once. ⇒ "Output data select" is displayed.
Press once, and present set value blinks.
Input the setting value with.
(Initial value:0)
Press to validate it.
Calculated value
(hold synchronized)
Calculated value
(hold not synchronized)
Chapter
9
OPTION INTERFACE
66
3. Presstwice to go back to the indicated value display.
■Signal timing
● MinusTurns on when the output indicated value data is negative.
● OVERTurns on at over-scale time.
-LOAD、+LOAD、OFL1、OFL2
● STABTurns on at stable-time. See "5-7. Motion Detect" on page 41.
● STROBEStrobe pulses are output in synchronization with BCD data.
For reading the data, use the rising edges of the pulses when the
logic is negative, and the falling edges of the pulses when the logic
is positive.
66
■Logic switching
BCD data
OVER
Strobe range
STROBE
0
1
0
1
STAB
(duty 50%)
Update
BCD data
OFF
ON
BCD
1cycle
1cycle(max)
Updating is stopped.
0
1
STROBE
data hold
9 OPTION INTERFACE
Select the logic of the BCD data output signal. Make selection with B14.
When COM and B14 are open:Negative logic
When COM and B14 are short-circuited :Positive logic
Read at least 2 cycles after inputting the logic switching.
■BCD data hold
Updating of the BCD data output signal is stopped. Also, the STROBE output turns OFF.
Make selection with A14.
When COM and A14 are open:Update stop canceled
When COM and A14 are short-circuited :Update stopped
Chapter
9
67
Read at least 2 cycles after inputting the BCD data hold.
As long as the BCD data hold is input, data can be read reliably, but because the data is not updated,
OPTION INTERFACE
it may be different from original data.
67
9 OPTION INTERFACE
Even if this setting is changed, A/D conversions speed does not change.
Make the output rate slower when reading fails due to a low processing
performance of the device that should receive BCD.
Point
Ch 2
H
I
LO
Ch
1
H
I
LO
FNC
H
I
Ch
1
Ch 2
OK
LO
PEAK
HOLD
H
I
LO
H
I
LO
ESC
■Output rate
《Setting value》
0:104:300
1:305:500
2:506:1000
3:100【times/sec.】
* When the A/D conversion speed is 300 times/sec., settings are available up to 4: 300
[times/sec.].
◇Output rate setting method
Chapter
9
OPTION INTERFACE
1. Select setting mode 5.
Press→four times.
2. Set "Output rate."
Press Twice. ⇒ "Output rate" is displayed.
Press , and present set value blinks.
Input the setting value with .
(Initial value:3)
Press to validate it.
3. Press twice to go back to the indicated value display.
■BCD/binary select
The output data selection can be changed to binary.
《Setting value》
0:BCD
1:Binary
68
68
◇BCD/binary setting method
Ch 2
HILO
Ch
1
HILO
FNC
H
I
Ch
1
Ch 2
OK
LO
PEAK
HOLD
H
I
LO
H
I
LO
ESC
1. Select setting mode 5.
Press→four times.
2. Select "BCD/binary select."
Press three times. ⇒ "BCD/binary select" is
displayed.
Press , and present set value blinks.
Input the setting value with .
(Initial value:0)
Press to validate it.
3. Press twice to go back to the indicated value display.
9 OPTION INTERFACE
■Pin assignment at binary
A1
A2
A3
A4
A5
A6
A7
A8
A9
A10
A11
A12
*
OUT
OUT
OUT
OUT
OUT
OUT
OUT
OUT
OUT
OUT
OUT
COM
0
2
1
2
2
2
3
2
4
2
5
2
6
2
7
2
8
2
9
2
10
2
B1
B2
B3
B4
B5
B6
B7
B8
B9
B10
B11
B12
*
OUT
OUT
OUT
OUT
OUT
OUT
OUT
OUT
OUT
OUT
OUT
COM
12
2
13
2
14
2
15
2
16
2
Data Switching 0
(Response)
Data Switching 1
(Response)
Minus (Polarity)
OVER
STAB
Chapter
9
A13
A14
A15
A16
OUT
IN
IN
-
Data Switching 0B15
11
2
B13
Data HoldB14
B16
OUT
IN
IN
-
STROBE
Logic Switching
Data Switching1
OPTION INTERFACE
* With the B8 and B9 outputs, the data specified by the output data selection or
by A15 and B15 (data select) can be checked.
69
69
9 OPTION INTERFACE
9-2.RS-232C interface
The RS-232C is an interface to read the indicated value and status of the F350 and to write
parameters into the F350.
This interface is convenient to process controls, totals, records, etc., by connecting the F350 to a
computer, process controller, sequencer or the like.
Character length: 7 or 8 bit Selectable
Stop bit:1 or 2 bit Selectable
Parity :Parity none,odd or even Selectable
Delimiter:CR, CR+LFSelectable
Chapter
9
OPTION INTERFACE
Code:ASCII
■Connector pin assignment
1
2
3
4
5
6
7
8
9
Case
Adaptable connector 9-pin D-SUB connector
-
IN
OUT
OUT
*
-
OUT
IN
-
*
Connector: XM3D-0921(Omron) or equivalent
Cover:XM2S-0913(with inch screw thread #4-40)(Omron) or equivalent
(connector & cover optional type: CN34)
RXD
TXD
DTR
GND
RTS
CTS
SHIELD
70
70
■Connection
Ch 2
H
I
LO
Ch
1
H
I
LO
FNC
H
I
Ch
1
Ch 2
OK
LO
PEAK
HOLD
H
I
LO
H
I
LO
ESC
9 OPTION INTERFACE
Example)F350
XM3D-0921 or equivalent
PC etc.
D-sub9pin
11CD
2RXD2RXD
3TXD3TXD
4DTR4DTR
5GND5GND
66DSR
7RTS7RTS
8CTS8CTS
99RI
* This connection diagram shows cabling applicable to the case where your PC is DTE (data
terminal equipment). For connection with DCE (data circuit-terminating equipment), such as a
modem, use straight type cabling.
* Prepare cabling after reconfirmation of the connector shape and signal lines (pin assignments) of
the equipment you use.
■Communication mode
《Setting value》
0:Send/receive
This mode performs communication by a command from the host computer.
In this mode, you can read out the indicated value, status, set values and write in
set values.
1:Continuous send
This mode continuously transmits the indicated values and the status.
2:Print send mode
When the indicated value is printed (a print command is output to the SI/F), the
indicated value and status are sent from the F350 to host.
◇Communication mode setting method
1. Select setting mode 5.
Press→four times.
2. Set "Communication mode."
Press once. ⇒ "Communication mode" is
displayed.
Chapter
9
Press , and present set value blinks.
Input the setting value with .(Initial value:0)
OPTION INTERFACE
Press to validate it.
3. Press twice to go back to the indicated value display.
71
71
9 OPTION INTERFACE
Ch 2
H
I
LO
Ch
1
H
I
LO
FNC
H
I
Ch
1
Ch 2
OK
LO
PEAK
HOLD
H
I
LO
H
I
LO
Baud rate
0:9600 bps 2:38400 bps
111 00
Flow control
0: None 1:RTS/CTS
Character length
0:7bit 1:8bit
Parity bit
0: None 1: Even 2: Odd
Stop bit
0:1bit 1:2bit
1:19200 bps 3:57600 bps
ESC
Ch 2
H
I
LO
Ch
1
H
I
LO
FNC
H
I
Ch
1
Ch 2
OK
LO
PEAK
HOLD
H
I
LO
H
I
LO
ESC
■Communication conditions
Set the communication conditions according to the host.
◇Communication conditions setting method
1. Select setting mode 5.
Press→four times.
2. Set "Communication conditions."
Press twice. ⇒ "Communication
conditions" is displayed.
Press , and present set value blinks.
Input the setting value with .
(Initial value:11100)
Move to the lower digits with .
Press to validate it.
Chapter
9
OPTION INTERFACE
3. Press twice to go back to the indicated
value display.
■Delimiter
Set the delimiter for sending messages from the F350.
《Setting value》
0:CR
1:CR+LF
◇Delimiter setting method
1. Select setting mode 5.
Press→four times.
72
72
2. Set "Delimiter."
Press three times. ⇒ "Delimiter" is displayed.
Press , and present set value blinks.
Input the setting value with .
(Initial value:11100)
Press to validate it.
3. Press twice to go back to the indicated value display.
■Communication format
R
.321+AR
CRA
54
Delimiter
Host
F350
Sign, decimal point, 5-digit indicated value.
R
.321+BR
CRB
54
Delimiter
Host
F350
Sign, decimal point, 5-digit indicated value.
R
.321+CR
CRC
54
Delimiter
Host
F350
Sign, decimal point, 5-digit indicated value.
R
00000DR
CRD
00
Delimiter
Host
F350
Hold
Stable
Near Zero
HI
OK
LO
0:OFF 1:ON
0:OFF 1:ON
0:OFF 1:ON
0:OFF 1:ON
0:OFF 1:ON
0:OFF 1:ON
Undefined
(Calculated value)
(Calculated value)
(Calculated value)
R
00000ER
CRE
00
Delimiter
Host
F350
HI(ch2)
0:OFF 1:ON
0:OFF 1:ON
0:OFF 1:ON
0:OFF 1:ON
0:OFF 1:ON
0:OFF 1:ON
Undefined
LO(ch1)
HI(ch1)
ALM(ch1)
LO(ch2)
ALM(ch2)
1.Send/receive [Reading the indicated value/status]
Function
Indicated value
(Calculated
value)
Header
MainSub
RA
9 OPTION INTERFACE
Reading the indicated value/status
Protocol
Indicated value
(ch1)
Indicated value
(ch2)
Status1
(Calculated
value)
RB
RC
RD
Chapter
73
Status2
(ch1,ch2)
RE
9
OPTION INTERFACE
73
9 OPTION INTERFACE
R
00000FR
CRF
00
Delimiter
Host
F350
OFL2(ch1)
0:OFF 1:ON
0:OFF 1:ON
0:OFF 1:ON
0:OFF 1:ON
0:OFF 1:ON
Undefined
Undefined
+LOAD(ch1)
-LOAD(ch1)
OFL1(ch1)
cErr(ch1)
R
00000GR
CRG
00
Delimiter
Host
F350
OFL2(ch2)
0:OFF 1:ON
0:OFF 1:ON
0:OFF 1:ON
0:OFF 1:ON
0:OFF 1:ON
Undefined
Undefined
+LOAD(ch2)
-LOAD(ch2)
OFL1(ch2)
cErr(ch2)
R
.321+JR
CRJ
54
Delimiter
Host
F350
Sign, decimal point, 5-digit indicated value.
Function
Header
MainSub
Reading the indicated value/status
Protocol
Status 3
(ch1)
Status 4
(ch2)
RF
RG
Chapter
9
OPTION INTERFACE
Indicated value
(Real time
Calculated
value)
74
74
RJ
2. Send/receive [Setting value writing/reading]
10000000WCR
Host
F350
* For a setting value with ch switching, make sure to write in advance the
command
1:ch1 2:ch2
Setting values without channel switching are disregarded.
channel number selected, and then write or read the setting value.
The command is fixed at 00.
0 fixation
00 fixation
4321+11WCR5
Host
F350
Sign, decimal point, 5-digit indicated value.command
* When the number of digits of the setting value is smaller than five, put a zero in the higher digit(s).
Example) 3 digits
2100+**WCR3
00000**WCR3
* For Motion Detect and Zero Tracking, put a zero in the 3rd digit.
10510**WCR0
Example)Time 1.5sec.
Range 10
(For a setting value with a decimal point, simply write
* When the setting value is unsigned, put a zero in the sign digit.
the numerical value excluding the decimal point.)
W
4321+11
11
5
Delimiter
Host
F350
CR
W
Sign, decimal point, 5-digit indicated value.command
(A setting value with a decimal point results in the numerical value excluding the decimal point.)
* The mode number of the setting value of HI/LO limit is "0."
* For setting mode numbers and setting item numbers, see "11-1.
The list of initial setting value" on page 88.
1 1
Setting value item number
Setting mode number (F*)
●"Setting protect" and "key protect" can be read but cannot be written.
●"Initialize" and "password" cannot be either read or written.
●When the protect of the setting value you are going to write is ON, it cannot be written. Turn OFF
the protect by key operation.
●If an output value is written in command 42 (zero calibration), simulated zero calibration is
executed with the written value. Also, if an output value is written in command 43 (equivalent input
calibration), equivalent input calibration is executed.
Point
Channel selection (Setting value with channel switching)
Function
Header
MainSub
9 OPTION INTERFACE
Protocol
channel
switching
Function
Various
settings
Function
w
MainSub
w
MainSub
-
Header
-
Header
Setting value writing
Protocol
Setting value reading
Protocol
Various
settings
w
-
* Command definition
75
Chapter
9
OPTION INTERFACE
75
9 OPTION INTERFACE
CCRZ
Host
F350
CCRS
Host
F350
CCRE
Host
F350
CCRF
Host
F350
CCRG
Host
F350
CCRH
Host
F350
CCRI
Host
F350
CCR1
Host
F350
CCR2
Host
F350
CCR3
Host
F350
CCR4
Host
F350
●Give commands with channel switching after setting the channel by ch switching.
For details, see " 1.Send/receive [Reading the indicated value/status]" on page 73.
●If the calibration protect is ON, zero calibration and span calibration cannot be
performed.
Point
3. Send/receive[Command]
Function
Zero calibration
(with channel
switching)
Equivalent input
calibration
(with channel
switching)
Digital Zero
(simultaneous on both
channels)
Digital Zero reset
(simultaneous on both
channels)
Digital Zero
(with channel
switching)
Header
MainSub
CZ
CS
CE
CF
CG
Movement directive
Protocol
Chapter
9
OPTION INTERFACE
Digital Zero reset
(with channel
CH
switching)
SI/F print commandCI
SECTION
ON
SECTION
OFF
T/H
ON
T/H
OFF
C1
C2
C3
C4
76
76
4. Send/receive [Returning an error]
E*R
Host
F350
Delimiter
1: The command or message length is improper.
2: The protect of the setting value you are going to write
3: The format (whether or not to sign, the number of
4: Other communication errors (parity error, frame error, etc.)
value you are going to write is improper.
calibration protect is ON.
digits, other than the numerical value) of the setting
is ON, or a calibration command is received while the
50+,68.47.1+1,32
1098765432121101
S8S7S6S5S4S3S2S1,,S9
F350
G S
Delimiter54.32+1
ch2
Calculated valuech1
S1S2S3S4S5
S:Stab le
M:Not Stable
H:Calculated value
Hold
O:ch1
Over Load
(±LOAD、OFL1,2)
A : ch1 AlarmC:ch1 Calibration
error
N:No error
O:ch2
Over Load
(±LOAD、OFL1,2)
A : ch2 AlarmC:ch2 Calibration
error
N:No error
H:Calculated value
HI ON
L:Calculated value
LO ON
G:Calculated value
HI/LO OFF
N:Calculated value
HI/LO ON
F:Calculated
value Comparison
OFF
H:ch1 HI ON
L:ch1 LO ON
G:ch1 HI/LOOFF
N:ch1 HI/LOON
H>(SorM)O > A > C > NO> A > C > NN> (HorL) > F > GN> (HorL) > G
S6S7S8S9
H:ch2 HI ON
L:ch2 LO ON
G:ch2 HI/LO OFF
N:ch2 HI/LO ON
N:Near Zero OFF
Z:Near Zero ON
A:ch1
Zero Tracking OFF
T:ch1
Zero Tracking ON
A:ch2
Zero Tracking OFF
T:ch2
Zero Tracking ON
N >(HorL)> G
232221201918171615252413 14282726
3736353433323129 30
Returning an error
Function
Header
MainSub
9 OPTION INTERFACE
Protocol
Error occurrence
(send/receive mode)
ER
■Continuousness / print send mode
Continuousness / print send
Function
Header
MainSub
Protocol
Indicated
value
&
Status
GS
77
Chapter
9
OPTION INTERFACE
77
9 OPTION INTERFACE
+
VOUT
+
IOUT
-
GND
-
GND
DAV (D/A converter voltage output)
DAI (D/A converter current output)
9-3.D/A converter output
A D/A converter is provided for obtaining analog output synchronized with the indicated value of
the F350.The analog output ranges are -10 - +10V output and 4 - 20mA constant-current output.
By using the D/A zero setting and D/A full scale setting functions, analog output can be obtained
between zero (0V, 4mA) and full scale (10V, 20mA) with respect to the predetermined digital value.
The output circuit and the main unit circuit are isolated.
The resolution is 1/10000 with respect to the ranges of -10 to +10V and of 4 to 20mA, and the
conversion speed follows the A/D conversion speed.
The maximum voltage output range are about ±11V, the maximum current output range are about
2.4 - 21.6 mA
■Connection
Chapter
For the output terminals of the D/A converter option, a cage clamp type terminal block is used.
Make connections by the following procedure.
9
OPTION INTERFACE
㧗
㧙
㧰㧛㧭
㧯㧻㧺㨂㧱㧾㨀㧱㧾
㨂ޓ㧻㨁㨀
ޓ
78
78
Connecting to cage clamp terminal block
5~6mm
1.
Strip the casing 0.2in (6mm) on the cable to be
connected.
Twist the bare wire to fit the terminal hole.
2. Insert the supplied screwdriver into the upper
hole and lift upward.
3. Insert the twisted wires into the lower hole.
4. Pull the screwdriver out from the upper hole.
5. Make sure cable is clamped securely and does
not come out with a slight tug.
9OPTION INTERFACE
● Cable can be from 24 ~ 14AWG (0.2 ~ 2.5mm2).
● It is not necessary to solder the cable wires or to fix a solderless terminal.
● If several cables to be inserted to the same hole, twist those cable wires together
and insert.
● Connect the cable with the power supply of the main body turned off without fail.
■Output data select
《Setting value》
0:Calculated valueOutput is made in synchronization with the calculated value.
1:ch1 valueOutput is made in synchronization with the ch1 value.
2:ch2 valueOutput is made in synchronization with the ch2 value.
3:Zero fixOutput is fixed at the zero output (0V for voltage output, 4mA
CAUTION
Chapter
9
for current output). After fixing, you will go to the output
adjustment mode, so that zero adjustment can be made.
See "■ Adjustment by fixed output" on page 81.
79
4:Full scale fixOutput is fixed at the full scale (10V for voltage output, 20mA
for current output). After fixing, you will go to the output
adjustment mode, so that gain adjustment can be made.
See "■ Adjustment by fixed output" on page 81.
79
OPTION INTERFACE
9 OPTION INTERFACE
Ch 2
HILO
Ch
1
HILO
FNC
H
I
Ch
1
Ch 2
OK
LO
PEAK
HOLD
H
I
LO
H
I
LO
ESC
Ch 2
H
I
LO
Ch
1
HILO
FNC
H
I
Ch
1
Ch 2
OK
LO
PEAK
HOLD
H
I
LO
H
I
LO
◇ Output data select setting method
1. Select setting mode 5.
Press→four times.
2. Set "Output data select."
Press once. ⇒ "Output data select" is
displayed.
Press , and present set value blinks.
Input the setting value with .
(Initial value:0)
Press to validate it.
3. Press twice to go back to the indicated value display.
■Zero/Full scale setting
Chapter
9
OPTION INTERFACE
Zero scaleSet the indicated value to output 0V for voltage output, and 4mA for
current output.
Full scaleSet the indicated value to output 10V for voltage output, and 20mA for
current output.
《Setting value》
Zero/Full scale value-99999~99999
* Set so that the Zero scale value < Full scale value
.
◇ Zero/Full scale setting method
1. Select setting mode 5.
Press→four times.
2. Set "Zero scale value," and "Full scale value."
① Set "Zero scale value."
Press twice. ⇒ "Zero scale value" is
displayed.
Press , and the highest digit blinks.
Input the setting value with , and move to the lower digit with .
* For inputting a negative sign, press
.
Press to validate it.
80
80
9 OPTION INTERFACE
H
I
Ch
1
Ch 2
OK
LO
PEAK
HOLD
H
I
LO
HILO
ESC
H
I
Ch
1
Ch 2
OK
LO
PEAK
HOLD
H
I
LO
H
I
LO
H
I
Ch
1
Ch 2
OK
LO
PEAK
HOLD
H
I
LO
H
I
LO
Adjust with the keys.
Zero scale adjusted value
H
I
Ch
1
Ch 2
OK
LO
PEAK
HOLD
H
I
LO
H
I
LO
H
I
Ch
1
Ch 2
OK
LO
PEAK
HOLD
H
I
LO
H
I
LO
Full scale adjusted value
Adjust with the keys.
ZERO
ESC
② Set "Full scale value."
Press once. ⇒ "Full scale value" is
displayed.
Press , and the highest digit blinks.
Input the setting value with , and move to the lower digit with.
* For inputting a negative sign, press
.
Press to validate it.
3. Press twice to go back to the indicated value display.
■Adjustment by fixed output
Make output adjustment on the adjustment screen appearing after selecting each fixed output by
D/A output data select setting and pressing .
While monitoring the output value, make fine adjustment of the output with , and
press to validate it, so that the adjusted value (%) is registered.
◇Zero fixed output
◇Full fixed output
Chapter
9
* Keeping on pressing results in continuous operation.
* The adjustable range of voltage output is approximately ±2.0V, and that of current output is
approximately ±1.6mA.
* Press to bring the adjusted value back to the central point (0%).
* Press to stop adjustment and go back to the indicated value display screen.
* By initializing the system settings, the adjusted value is cleared to go back to 0%.
81
81
OPTION INTERFACE
10 SPECIFICATIONS
10 SPECIFICATIONS
10-1. Specifications
■Analog section
Number of
sensor inputs2 inputs
Sensor excitationDC10V、2.5V±5%(It depends on the setting.)
Output current:Within 60mA (2 channels total)
* The following are common to each input.
Signal input range-3.0~+3.0mV/V
AccuracyNon-linearity: Within 0.02%FS±1 digit (at a 3.0mV/V input)
Zero drift:Within 0.5μV/ ℃RTI
Gain drift:Within 0.01%/ ℃
Chapter
10
SPECIFICATIONS
A/D converterRate:Selectable from 3000 times/sec and 300 times/sec
Resolution:24 bits (binary), 1/10000 with respect to 1mV/V
Voltage outputApproximately 2V per load cell input of 1mV/V,
load resistance 2kΩ or more
■Display section
Display7-segment LED, LED lamps
Main display: Character height 15mm,
6-digit, indicated value -99999 to 99999
Sub display:Character height 8mm,
5-digit, indicated value -19999 to 19999
(At the highest digit of the display, or nothing is displayed.)
Status display: HI、OK、LO、PEAK、HOLD
Red LED
Decimal point: 0、0.0、0.00、0.000
(Calculated values and ch1/ch2 values are all synchronized.)
Display itemDisplay frequency:Selectable from 3, 6, 13, and 25 times/sec.
■Setting section
Setting methodSetting by inputting with membrane keys
Setting valueNOVRAM (non-volatile RAM) or
lithium-battery-backed-up C-MOS RAM
82
82
10 SPECIFICATIONS
■External input/output section
External inputDry contact input circuit (minus common type) for hold control, etc.
(internal circuit power supply voltage 12V)
External outputOpen collector output circuit (sink type) for comparison output, etc.
Rated voltage: 30V, driving current: 30mA or less
■Interface
SIF2-wire serial interface for connecting a UNIPULSE-manufactured printer,
external display, etc.
Transmitting method: Asynchronous
Baud rate:600bps
Output data:Allocations to the GROSS area and NET area are
selected by the setting of "calculated value & ch1
value," "calculated value & ch2 value," or "ch1
value & ch2 value."
■Options
BCD parallel data output (BCD)
RS-232C interface (232)
D/A converter voltage output (DAV)
Open collector output circuit (sink type)
Data format:BCD or binary selectable
Output rate:10、30、50、100、300、500、1000 times/sec.
Rated voltage:30V
Driving current:30mA or less
Output data:Selected by the setting of "calculated
value," "ch1 value," or "ch2 value," or
externally.
Transmitting method:Asynchronous
Baud rate:9600bps to 57600bps
Flow control:RTS/CTS controllable
Output range:±10V
Load resistance:2kΩ or more
D/A conversion speed: Follows the original A/D conversion speed
Resolution:1/10000
Zero/span adjustment range:±10%F. S.
Non-linearity:±0.05%F.S.
Output data:Selected by the setting of "calculated
Output range:4 to 20mA
Load resistance:350Ω or less
D/A conversion speed: Follows the original A/D conversion speed
Resolution:1/10000
Zero/span adjustment range:±10%F. S.
Non-linearity:±0.05%F.S.
Output data: Selected by the setting of "calculated