Unipulse F350 Operation Manual

F350
OPERATION MANUAL
01MAR2015REV.1.04
Introduction
Applications

Introduction

Introduction
The F350 is an indicator for various strain gauge sensors that measure pressure, load, torque, etc. Equipped
with ultrahigh-speed peek hold and HI/LO limit comparison functions, the F350 is for general purpose use,
such as an automatic assembly machine and testing machine.
To take full advantage of high performance of F350, thoroughly read this operating manual first before use
and understand the explanations contained herein for correct operating procedures.Also, carefully store
this instruction manual so that it can be referred to at any time.
I
I
Safety Precautions
WARNING
CAUTION
This sign forewarns the presence of hazards that could result in serious injury or fatality when incorrectly handled.
Warning on design

Safety Precautions

Safety Precautions
BE SURE TO READ FOR SAFETY
Installation, maintenance and inspection of the F350 should be performed by personnel having technical
knowledge of electricity.
In order to have an F350 Weighing Controller used safely, notes I would like you to surely follow divide
into " " and " ", and are indicated by the following documents.
Notes indicated here are the serious contents related to safely.
Please use F350 after understanding the contents well.
This sign forewarns the presence of hazards that could result in serious injury
or fatality when incorrectly handled.
WARNING
CAUTION
This sign forewarns the presence of hazards that could result in personnel
injury or property damage when incorrectly handled.
WARNING   
For the entire system to function safely when the F350 becomes faulty or malfunctions,
provide a safety circuit outside the F350.
Since the F350 has no power switch, install a breaker.
Before using the F350 as described below, make sure to consult with our sales personnel.
- Use in environments not described in the operation manual.
- Use greatly impacting human lives and assets, such as medical devices, transport devices
entertainment devices, and safety devices.
II
II
WARNING   
This sign forewarns the presence of hazards that could result in serious injury or fatality when incorrectly handled.
Warning on installation
Warning on wiring
Warning during startup and maintenance
Do not disassemble, repair, or modify the F350. Doing so may cause a fire or an electric
Do not install in the following environments.
Do not connect a commercial power source directly to the signal input/output terminals.
Be sure to ground the protective ground terminal.
The attached AC cable is designed for domestic use in Japan, and its rating is 125V AC,
WARNING   
WARNING   
shock.
- Places containing corrosive gas or flammable gas.
- Where the product may be splashed with water, oil or chemicals.
10A. For use at voltages exceeding the rating and for overseas use, have a separate AC
cable prepared.
Before performing the following, make sure that no power is applied.
- Removal and installation of optional connectors and so forth;
- Wiring and connection of cables to a power input terminal;
- Wiring and connection of cables to a signal I/O terminal;
- Connection to protective grounding terminals.
For connection to the signal input/output terminals, check the signal names and pin
assignment numbers, and then carry out wiring properly.
After wiring, be sure to mount the attached terminal block cover. Otherwise, it may cause an
electric shock.
To take measures against lightning surge, install a lightning surge protector (optionally
available).
Do not connect anything to unused terminal(s).
Before applying power, carefully check the wiring, etc.
Use a power supply voltage and load within the specified and rated ranges.
Do not damage the power cord. Doing so may cause fire or electric shocks.
Do not touch any signal input/output terminal while applying power. Doing so may cause
electric shocks or malfunctions.
If the cover of the main body is opened, it may cause an electric shock internally. Even if the
power is off, the internal capacitor is charged. Contact us for internal inspection or repair.
In the case of smoke, an abnormal smell or strange sound, immediately turn off the power,
and disconnect the power cable.
III
III
CAUTION 
This sign forewarns the presence of hazards that could result in personnel injury or property damage when incorrectly handled.
Caution on installation
Caution on wiring
Caution during startup and maintenance
CAUTION 
CAUTION 
Use the F350 as it is incorporated in a control panel, etc.
Do not install in the following environments.
- Locations where temperature or humidity exceeds specifications;
- Locations subjected to drastic temperature fluctuations or icing and condensing;
- Outdoors or locations above 2,000m;
- Locations exposed to direct sunlight;
- Locations subject to dust accumulation;
- Locations with poor ventilation;
- Locations with a lot of salt and metal powder;
- Locations where the main unit is subject to direct vibration and shock.
Take adequate shielding measures when using at the following locations.
- Near a power line.
- Where a strong electric field or magnetic field is formed.
- Where static electricity, relay noise or the like is generated.
Install the F350 as far away from devices generating high frequency, high voltage, large
current, surge, etc., as possible. Also, carry out wiring separately from their power lines. Do
not carry out parallel wiring and common wiring.
Do not use it, broken down.
Tighten the screws for the power input terminal at the specified torque.
If they are loose, shorts, fire or malfunctions may occur.
Tightening torque: 0.5N •m
For sensors, external inputs/outputs and options, use shielded cables.
The temporary overvoltage applied to the power should not exceed 1500V.
For turning on/off the power, be sure to keep intervals of 5 seconds or more.
After power-on, make sure to warm up the F350 for at least 30 minutes or more before use.
If the F350 is not used by the specified method, its protective performance may be impaired.
Maintenance
- When performing maintenance, disconnect the power.
- Do not wipe with a wet rag, or with benzine, thinner, alcohol, etc. Doing so may cause
discoloration or deformation of the F350. In the case of heavy contamination, wipe off
the contamination with a cloth after dipping it into a diluted neutral detergent and wringing
it well, and then wipe with a soft, dry cloth.
IV
IV
CAUTION 
This sign forewarns the presence of hazards that could result in personnel injury or property damage when incorrectly handled.
Caution during transportation
Caution during disposal
This sign forewarns the presence of hazards that could result in serious injury or fatality when incorrectly handled.
CAUTION 
CAUTION 
When the F350 is shipped, spacers made of corrugated cardboard are used as cushioning
materials.
Though it is factory-designed so that shocks can sufficiently be absorbed, breakage may
result if shocks are applied when the spacers are reused for transportation. If you send the
F350 to us for repair, etc., take adequate measures against shocks by using polyurethane
materials, etc., separately.
If you dispose of the product, handle it as industrial waste.

About the Built-in Lithium Battery

WARNING   
Never disassemble, deform under pressure or throw the battery
into fire.The battery may explode, catch fire or leak.
V
V
Statement of Conformation to EC Directives
        )
Overvoltage category II Pollution degree 2
The combination of the main unit of the F350 and a lightning surge protector
conforms to EN61000-4-5 (lightning surge immunity) in the EMC Directives.
Refer to " Connection of a Lightning Surge Protector (Power Cable)" on page VII
for information regarding lightning surge protector connection.
Point

Statement of Conformation to EC Directives

Statement of Conformation to EC Directives
* The F350 is a CE-compliant product. For use, observe the following.
The F350 Digital Indicator is a CE-marked EC-Directive-conforming product
(by the Council of the European Union).
- Low Voltage Directive EN61010-1:2010
EN62311:2008 (test distance: 0cm)
- EMC Directive EN61326-1:2006
EN55011:2009, A1:2010 Group1, ClassA
EN61000-3-2:2006, A1:2009, A2:2009
EN61000-3-3:2008
EN61000-4-2:2009
EN61000-4-3:2006, A1:2008, A2:2010
EN61000-4-4:2004, A1:2010
EN61000-4-5:2006
EN61000-4-6:2009
EN61000-4-8:2010
EN61000-4-11:2004
When installing, attention should be given to the following.
1. Since the F350 is defined as open type (built-in equipment), be sure to use it as installed
and fixed to a panel, etc.
2. The power cable attached to this product as standard can be used with 100V AC power in
Japan.(Nominal rated voltage: 125V AC)
For using this product in a country outside Japan, use a power cable certified in that
country.
3. For cables other than the power cable (sensor, external input/output, option), use shielded
cables.
Also, mount the attached ferrite core to the strain gauge sensor cable.
VI
VI
Connection of a Lightning Surge Protector (Power Cable)
F350
Lightning surge protector
MAINTRAB MNT-1D is a registered trademark of
PHOENIX CONTACT.
An EU-outlet-shape AC power cable is required.
CA325AC3P-CEE7/7-B2: CEE7/7 plug cable (2m)

Connection of a Lightning Surge Protector (Power Cable)

Connection of a Lightning Surge Protector (Power Cable)
Connect a lightning surge protector against lightning surge.
EMC Directive EN61000-4-5 (Lightning Surge Immunity) is met by the F350 body in combination with a
lightning surge protector.
Lightning surge protector [MAINTRAB MNT-1D]
The lightning surge protector [MAINTRAB MNT-1D] is not included as standard.
In addition, as optional parts, we sell EU-outlet-shape 250V AC withstanding cables (following illustration:
European standard) and lightning surge protectors in pairs (TSU02). For details, contact our sales
department.

RoHS-Compliant Product

The parts and attachments (including the instruction manual, packaging box, etc.) used for this unit are
compliant with the RoHS Directive restricting the use of hazardous substances with regard to adverse
effects on the environment and human body.
What is RoHS?
It is an abbreviation for Restriction on Hazardous Substances, which is implemented by the European
Union (EU). The Directive restricts the use of six specific substances in electric and electronic equipment
handled within EU borders. The six substances are lead, mercury, cadmium, hexavalent chromium, PBB
(polybrominated biphenyls), and PBDE (polybrominated diphenyl ethers).
VII
VII
CONTENTS

CONTENTS

CONTENTS
1 OUTLINE . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . 1
1-1. Main features of the F350. . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . 1
1-2. Contents of the package. . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . 2
1-3. About connectable devices. . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . 3
1-4. Appearance description . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . 3
2 INSTALLATION & CONNECTION. . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . 7
Front panel . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . 3
Rear panel. . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . 6
2-1. Installation . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . 8
2-2. Connection . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . .9
Power input terminal connection . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . 9
Analog input/output connector connection . . . . . . . . . . . . . . . . . . . . . . . . . . . . 10
How to remove the terminal block . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . 10
How to connect and install the terminal block . . . . . . . . . . . . . . . . . . . . . . . . . . 11
Attachment of a ferrite core . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . 11
Strain gauge sensor connection . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . 12
Strain gauge sensor cable coloration . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . 13
Voltage output connection. . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . 13
External I/O connection . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . 14
How to assemble the connector . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . 15
SI/F interface connection. . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . 15
3 SETTING MODE CONFIGURATION. . . . . . . . . . . . . . . . . . . . . . . . . . . . . . 16
3-1. Setting mode composition . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . 16
3-2. Key operation . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . 18
3-3. Calling a setting mode . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . 22
VIII
4 CALIBRATION PROCEDURE . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . 23
4-1. How to calibration . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . 23
Equivalent input calibration . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . 23
Actual load calibration . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . 23
4-2. Procedure of equivalent input calibration and calibration protect. . . . . . . . . . . . . . . . 24
VIII
CONTENTS
CONTENTS
5 EXPLANATION OF INDICATED-VALUE-RELATED FUNCTIONS . . . . . . . 35
5-1. A/D conversion speed select . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . 35
A/D conversion speed select setting method . . . . . . . . . . . . . . . . . . . . . . . . . . 35
5-2. Analog filter . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . .36
Analog filter setting method. . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . 36
5-3. Digital filter . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . 37
Digital filter setting method . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . 37
5-4. Digital Zero . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . .38
Digital Zero / Digital Zero reset by means of keys . . . . . . . . . . . . . . . . . . . . . . 38
Digital Zero by means of the external signal "D/Z input". . . . . . . . . . . . . . . . . . 39
5-5. D/Z synchronous mode . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . 39
D/Z synchronous mode setting method . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . 39
5-6. Zero Tracking . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . 40
Zero Tracking setting method . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . 40
5-7. Motion Detect . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . 41
Motion Detect setting method . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . 42
5-8. Display frequency . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . 43
Display frequency setting method . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . 43
6 EXPLANATION OF COMPARISON FUNCTIONS . . . . . . . . . . . . . . . . . . . . 44
6-1. Calculation mode . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . 44
Calculation mode setting method . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . 44
6-2. HI/LO limit comparison of calculated value / ch1 value / ch2 value . . . . . . . . . . . . . . 45
HI/LO limit setting method . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . 45
6-3. Hysteresis. . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . .45
Hysteresis setting method . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . 46
6-4. Near Zero . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . .47
Near Zero setting method . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . 47
6-5. Calculation comparison timing . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . 48
Calculation comparison timing setting method . . . . . . . . . . . . . . . . . . . . . . . . . 48
IX
6-6. Comparison OK output conditions . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . 49
Comparison OK output conditions setting method . . . . . . . . . . . . . . . . . . . . . . 49
7 EXPLANATION OF HOLD FUNCTIONS . . . . . . . . . . . . . . . . . . . . . . . . . . . 50
7-1. Hold functions of calculated values . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . 50
IX
CONTENTS
CONTENTS
Flow of hold measurement . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . 50
7-2. Setting of hold mode. . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . 51
Hold mode setting method . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . 51
Sample hold . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . 51
Peak hold . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . 52
Bottom hold . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . 52
Average hold . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . 53
Average sample number . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . 53
7-3. Setting of hold section . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . 54
Hold section setting method . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . 54
All section . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . 55
External signal. . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . 56
External signal + time . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . 57
7-4. Auto reset function . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . 58
Auto reset setting method . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . 59
8 STANDARD INTERFACE . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . 60
8-1. External I/O. . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . .60
Connector pin assignment. . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . 60
About outputs . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . 60
About inputs . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . 61
About the RUN signal . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . 61
8-2. SI/FSerial interface . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . 62
SI/F Indicated value format . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . 62
SI/F print . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . 62
9 OPTION INTERFACE . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . 64
9-1. BCD interface . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . 64
Connection . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . 64
Connector pin assignment. . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . 65
Output data select . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . 66
Signal timing . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . 66
Logic switching . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . 67
BCD data hold . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . 67
Output rate . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . 68
BCD/binary select . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . 68
Pin assignment at binary . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . 69
9-2. RS-232C interface . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . 70
Communication specifications. . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . 70
Connector pin assignment. . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . 70
Connection . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . 71
Communication mode . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . 71
Communication conditions . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . 72
Delimiter . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . 72
Communication format . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . 73
X
X
CONTENTS
CONTENTS
Continuousness / print send mode . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . 77
9-3. D/A converter output . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . 78
Connection . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . 78
Output data select . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . 79
Zero/Full scale setting . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . 80
Adjustment by fixed output . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . 81
10 SPECIFICATIONS . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . 82
10-1. Specifications . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . .82
Analog section. . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . 82
Display section . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . 82
Setting section. . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . 82
External input/output section. . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . 83
Interface . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . 83
Options . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . 83
General performance . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . 84
Attachment . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . 84
10-2.Dimensions. . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . .85
Standard equipment . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . 85
Equipped with BCD parallel data output interface option . . . . . . . . . . . . . . . . . 85
Equipped with RS-232C interface option . . . . . . . . . . . . . . . . . . . . . . . . . . . . . 86
Equipped with D/A converter option . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . 86
10-3. F350 block diagram. . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . 87
11SUPPLEMENTS . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . 88
11-1.The list of initial setting value . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . 88
11-2.Self-check function . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . 94
Self-check method . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . 94
11-3.Setting protect . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . .95
Setting protect setting method . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . 95
11-4.Key protect . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . .96
Key protect setting method . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . 96
11-5.Initialize . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . 97
Initialize setting method. . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . 97
XI
11-6.Password . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . .97
11-7.Error/message list . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . 97
Main numerical display section . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . 97
Sub numerical display section. . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . 98
11-8.Troubleshooting . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . 99
XI
CONTENTS
CONTENTS
11-9. About disposal of the lithium battery . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . 101
Removal of the lithium battery. . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . 101
XII
XII

1-1. Main features of the F350

1 OUTLINE
The F350 is a digital indicator that is capable of two-channel inputs of strain gauge
sensors.
Calculations, such as addition and subtraction, and comparison judgments are made
by using the input values of two channels. The F350 is also suitable for balance
measurement.
The F350 is equipped with comparison judgments on the input value of each channel,
and analog voltage output.
The A/D conversion speed of each channel and the digital processing speed for
calculations and judgments are as high as 3,000 times/sec. It can also be switched to
300 times/sec.

1 OUTLINE

Chapter
1
OUTLINE
The F350 is equipped with hold functions with respect to calculated values.
(sample, peak, bottom, and average holds)
The F350's compact size of DIN96×96 realizes space-saving.
By mounting an optional interface according to the purpose, ease of load control can
be further improved.
1
1
CAUTION
F350 body・・・1 F350 operation manual・・・1
AC input cord・・・1
Conversion plug for
Mini screwdriver for Connector for BCD parallel
(with BCD parallel data output option)
AC input code・・・1
(with D/A converter option)
Ferrite core・・・2
with unity band
External input/output
(with solderless terminal)
connector・・・1
terminal block connection・・・1
data output・・・1
1 OUTLINE
Chapter
1
OUTLINE

1-2. Contents of the package

The packaging box contains the following.
Be sure to check them before use.
CAUTION
The attached AC cable is intended for domestic use in Japan, and it is rated as 125V AC, 10A. If it
is used at voltages exceeding the rating or it is used overseas, have a suitable AC cable ready
separately.
2
2
CAUTION
Strain gauge type sensor
(Option)
D/A Converter
Valve, recorder
(Option)
(Option)
Printer Display Various converters
PC
* + .1 * + .1
* + . 11 -
* 1 . &
2 ' # -
%J
%J
(0% '5%
*+.1
<'41
*1.&

F
350
Main status display
Sub status display
Setting key pad
Main numerical display
Sub numerical display section
section
section
section
1 OUTLINE

1-3. About connectable devices

Chapter
1
OUTLINE

1-4. Appearance description

3

Front panel

3
CAUTION
In this operation manual, the display status is expressed as follows:
Lighting Status display section Black
Numerical display section White
Blinking Lighting state + Lines in a radial pattern
Being unlitStatus display section Gray
Numerical display section Nothing displayed
LO
H
I
Ch
1
Ch 2
OK
LO
PEAK
HOLD
H
I
LO
H
I
Lighting
Blinking
Being unlit
Chapter
1
OUTLINE
1 OUTLINE
Main status display section
The status with respect to calculated values is indicated.
HI Lights when the indicated value is larger than the calculation HI limit.
Indicates that the external output "calculation HI" is ON.
OK Lights when the indicated value is smaller than the calculation HI limit and larger
than the calculation LO limit.
Indicates that the external output "calculation OK" is ON.
LO Lights when the indicated value is smaller than the calculation LO limit.
Indicates that the external output "calculation LO" is ON.
PEAK Blinks during hold detection, and lights when a hold is confirmed.
However, it is always unlit in all section sample hold.
HOLD Lights during hold detection and when a hold is confirmed.
However, it blinks when hold waiting in all section sample hold.
Main numerical display section
The following display patterns exist.
Indicated valueDisplays calculation results during measurements.
Error display Displays an error message when an error occurs.
Setting valueDisplays setting values in setting mode.
Sub status display section
The status with respect to ch1/ch2 values is indicated.
HI Lights when the indicated value is larger than the ch1/ch2 HI limit.
Indicates that the external output "ch1/ch2 HI output" is ON.
LO Lights when the indicated value is smaller than the ch1/ch2 LO limit.
Indicates that the external output "ch1/ch2 LO output" is ON.
Sub numerical display section
The following display patterns exist.
Indicated valueDisplays ch1/ch2 values during measurements.
Error display Displays an error message when an error occurs.
Setting value Displays the mode number and item number at the ch1 display section in
setting mode; the currently-selected weight value may be displayed at the ch2
display section according to the setting item
4
4
CAUTION
FNC
ESC
HI/LO
ZERO
HOLD
Setting key pad
These are keys for commanding settings and operations.
Use this key to go to a setting mode state.
Use this key to cancel setting/execution.Use this key to go back from a setting mode state to an indication display state.
Use this key to take a shortcut from an indication display state to HI/LO limit setting.
Use this key to go from an indication display state to ch1 or ch2 Digital Zero ready.
Use this key to execute/release the calculated value hold in an indication display state.
Use this key to switch between ch1 setting and ch2 setting, etc., in a Digital Zero
ready state or setting item select state.
Use this key to set whether or not to sign a setting value, when available, in a setting
value input state.
1 OUTLINE
Chapter
1
OUTLINE
Use this key to select a trailing setting item number in a setting mode state.Use this key to select the previous item number in a setting item select state.Use this key to increment the numerical value by one in a setting value input state.
Use this key to select a leading setting item number in a setting mode state.Use this key to select the next item number in a setting item select state.Use this key to decrement the numerical value by one in a setting value input state.
Use this key to execute Digital Zero reset in a Digital Zero ready state.Use this key to increment the leading setting mode number by one in a setting mode
state.
Use this key to move the setting digit by one in the lower direction in a setting value
input state.
Use this key to execute Digital Zero in a Digital Zero ready state.Use this key to move to a setting value input state in a setting item select state.Use this key to validate the setting value in a setting value input state.
5
5
CAUTION
Options slot
Power input terminal
Protective ground
Signal input/output
Analog input/output connector
connector
Frame ground
Chapter
1
OUTLINE
1 OUTLINE

Rear panel

Analog input/output connector
Connect a strain gauge sensor or voltage output.
* For connection, see "
Analog input/output connector connection" on page 10.
Signal input/output connector
Connect external input/output signals or SI/F (serial interface).
* For connection, see "
External I/O connection" on page 14.
Power input terminal block
Connect an AC input cord.
The input power source is free in the range of 100 to 240V AC (-15%, +10%). The frequency is 50/
60Hz.
* For connection, see "
Power input terminal connection" on page 9.
Protective ground
Be sure to ground the protective ground terminal to prevent electric shocks.
Frame ground
6
6
Please ground the frame ground terminal to prevent failures due to static electricity.
(The frame and the frame ground terminal are conducted.)
It may be better to remove depending on the environment of the installation location.
Options slot
One option board can in stall in the option slot.
BCD parallel data output (BCO)RS-232C interface (232)D/A converter voltage output (DAV)D/A converter current output (DAI)

2 INSTALLATION & CONNECTION

2 I
NSTALLATION & CONNECTION
The following are precautions related to connection to the signal input/output terminal block.
The precautions described here are important for safety.
Make connections after properly understanding the description.
WARNING
Do not connect commercial power directly to the signal input/output terminals.
Make connection to the signal input/output terminals with no power applied
because you may receive an electric shock.
For connections to the signal input/output terminals, carry out wiring properly after
confirming the signal names and pin assignment numbers.
Use of the F350 is limited to category II specified by EN61010. Overvoltage applied
to the signal input/output terminals should not exceed the value defined to category
II.
After wiring, make sure to mount the attached terminal block cover. Otherwise, you
may receive an electric shock.
Turn on the power after carefully checking the wiring, etc.
Chapter
2
INSTALLATION & CONNECTION
Do not touch the signal input/output terminals while the power is on. Otherwise, you
may receive an electric shock, or a malfunction may occur.
The F350 conforms with EMC Directives as a product used in industrial
environments (class A). If it is used in living environments, radio disturbance may
occur. In that case, take suitable measures.
CAUTION
Tighten the terminal screws at specified torque. If the terminal screws are loose,
short-circuiting, fire, or a malfunction may occur.
Use shielded cables.
7
7
2 INSTALLATION & CONNECTION
92mm
Panel-cut
+1 0
Panel thickness
1.6 to 3.2mm
dimensions
92mm
+1 0

2-1. Installation

To install the F350 into a control panel, use the following procedure.
1. Make a hole in the panel according to
Chapter
2
INSTALLATION & CONNECTION
the panel-cut dimensions.
2. Remove the screws (two), and remove
the guide rails from both sides.
* Do not use other screws than those installed to
the F350 body.
3. Fit in the F350 from the front of the panel.
4. Install from the rear the guide rails having
been removed from both sides in Step 2,
and fix them with the screws (two).
8
8

2-2. Connection

Within 6mm
Black
White
Green/Yellow
Color description are for the attached AC input cord.

Power input terminal connection

Connect an AC input cord. The input power source is free in the range of 100 to 240V AC (-15%,
+10%).
The frequency is 50/60Hz.
Make connections to the terminal block with a solderless terminal within 6mm in diameter as
shown in the illustration so as not to let the tip of the cable spread out.
1. Make sure that no power is applied.
2. Remove the screws (two), and remove
the terminal block cover.
3. Remove the two screws(M3) at
the terminal block
.
2 INSTALLATION & CONNECTION
Chapter
2
4. Align the solderless terminals with
the screw holes, and then tighten
the screws.
L Black N White
5. Install the terminal block cover, and fix it with the screws (two).
6. Remove the screws(M4) of the protective ground, align the crimp contacts
with the screw holes, and then tighten the screws.
Green/ Yellow
The protective ground is internally connected with frame ground.
*
INSTALLATION & CONNECTION
9
9
Chapter
Loosen. (Turn
counterclockwise
.)
When installing the terminal block to the F350
body, check its orientation. (See the illustration at
the right.)
Attention
Insert side
Right side
2 INSTALLATION & CONNECTION

Analog input/output connector connection

Connect a strain gauge sensor or voltage output.
Make connections according to the silk-screen printing on the rear panel of the F350.
Pin Assignment
12 SHIELD
2
11 -MONITOR(ch2)
10 +MONITOR(ch2)
INSTALLATION & CONNECTION
9 -MONITOR(ch1)
8+MONITOR(ch1)
7 -SIG(ch2)
6+SIG(ch2)
5 -SIG(ch1)
4+SIG(ch1)
3 -EXC*
2 +EXC*
1SHIELD
Compatible connector: ETB42-12P(OSADA CO.,Ltd.) or equivalent (Connector optional type:CN81)
  * EXC is common to ch1 and ch2.
CAUTION
Before connecting a strain gauge sensor, make sure to check the setting of excitation
voltage. For the setting of excitation voltage, see "4-2 Excitation voltage setting" on page
25.

How to remove the terminal block

1.
Loosen the screws (two) with a screwdriver.
2. Remove the terminal block by giving it a
strong pull.
10
10
2 INSTALLATION & CONNECTION
5mm
Pin No.12
Pin No.1
Adaptable plug ETB42-12P
Tighten.
(Turn clockwise.)
attachment
(Turn counterclockwise.)
Loosen.
2KP0Q
2KP0Q
Tighten.
(Turn clockwise.)
Tighten.
(Turn clockwise.)
Sensor cables
F350
A cable is twisted around a ferrite core.(One roll)
Ferrite core

How to connect and install the terminal block

Strip 5mm of the covering of the wire to be
1.
connected.
The size of connectable wires is from 0.21 to
3.31mm
2
(AWG12 to 24).
2. Twist the tip to such an extent that it will not spread out.
3. Loosen the screw with a screwdriver to open
the connection hole.
*A Phillips screwdriver 3 to 3.5mm #1 in shaft
diameter is recommended. (Precision screwdriver,
etc.)
4. Insert the wire into the connection hole so as
not to let the tip spread out.
Chapter
2
5. Tighten the screw with the screwdriver.
*0.5Nm of tightening torque is recommended.
6. Lightly pull the wire to make sure that
it is securely clamped.
7. Insert the wire-connected plug into
the F350 body, and tighten the
screws (two).

Attachment of a ferrite core

It is necessary to twist a power supply cable and sensor cables, such as a load cell, around an
attached ferrite core.
INSTALLATION & CONNECTION
11
11
2 INSTALLATION & CONNECTION
-SIG
-EXC
+EXC
+SIG
SHIELD
+IN
+OUT
-IN
-OUT
5
4
3
2
1
SHIELD
+SIG
-SIG
+IN
+OUT
-IN
-OUT
F350
12
7
6
SHIELD
+SIG
-SIG
12
7
6
-SIG
-EXC
+EXC
+SIG
SHIELD
+
IN
+
OUT
-
IN
-
OUT
F350
+
S
-
S
5
4
3
2
1
+IN
+OUT
-IN
-OUT
+S
-S
Short-circuit +EXC with +S and -EXC with -S for connecting a 6-wire strain­gage sensor.

Strain gauge sensor connection

4-wire sensor
Chapter
2
INSTALLATION & CONNECTION
6-wire sensor
12
12
2 INSTALLATION & CONNECTION
External equipment
8 or 10
9 or 11
MONITOR1or 2
F350 Inside Outside
MONITOR1or 2
SIG
SIG
A/D
Analog filter

Strain gauge sensor cable coloration

Cable colors of sensors may differ from one manufacturer to another (it may even differ from one
model to another for some products). Refer to the sensor manual (or data sheet) and check signal
names and colors in order to connect the cables correctly.
CAUTION
The excitation voltage of the F350 is 2.5V/10V. If the maximum excitation voltage of
the sensor is under 2.5V/10V, heating or damage may result.

Voltage output connection

The voltage output terminals output analog voltage in proportion to each sensor signal input.
The voltage output is approximately 2V per sensor input of 1mV/V.
It is useful for observing or recording waveforms by connecting a recorder, etc.
Example) Connection of output equivalent circuit and external instrument.
Chapter
2
Since the output voltage is taken out in a step prior to A/D conversion of the sensor input signal, it
Since the ±MONITOR terminals are not insulated from the internal circuit, use two-core shielded
CAUTION
is not synchronized with indicated values processed digitally, such as Digital Zero and digital
filter. Output in synchronization with indicated values requires an optional D/A converter.
twisted pair wires for connection with external equipment, and carry out with as short a wiring as
possible.
Do not short-circuit. Doing so will cause a failure.
Do not apply voltage externally. Doing so will cause breakage.
INSTALLATION & CONNECTION
13
13
2 INSTALLATION & CONNECTION
Vceo=30V(max)
Output
Output COM1
Ic=30mA(max)
COM
Input
F350
Sink type
PLC etc.
Plus common connection
A8 A10 B2 B6
A1,A7 B1
DC24V
Minus common connection
+24V
0V
A2 A6
COM2
+12V
F350 Inside
IN
IN
The external element is required to withstand Ic=10mA.Leakage from the external element is required to be 100μA or below.
B8~12
B7
Open OFF Short ON
Push Switch
Toggle Switch Relay
TTL Open Collector (ON when IN is HI)

External I/O connection

Details of signal, see "8-1.External I/O" on page 60.
How to connect external output (Sink type)
Chapter
2
The external output circuit is operated through an open collector. A1(A7) COM1 is the common
terminal. The open collector output capacity is 30mA and the withstand voltage is up to 30V.
INSTALLATION & CONNECTION
Output Transistor Status
Output Data Tr
How to connect external input
OFF OFF
ON ON
For driving, have an external power source ready.
(24V DC is shown in the illustration.)
(minus common connection of no-voltage contact input type)
A signal is inputted to the signal input circuit by short-circuiting or opening the input terminal and
the COM2 terminal. Short-circuiting is effected by means of a contact (such as a relay or a switch)
or a noncontact (such as a transistor or an open-collector TTL)
14
14

How to assemble the connector

Connector
Pan-head machine screw M2×8 (short) (two)
Nut M2 (four)
Pan-head machine screw M2×10 (long) (two)
Washer (two)
Case (two)
Screw (two)
RNC[
SI/F
A11
Inside Outside
F350
A12
Set the connector and screws (two) into the grooves of the case (one side).
1.
2 INSTALLATION & CONNECTION
Chapter
2
2. Cover with the other case, and fit the cases.
3. Tighten the M2×8 pan-head machine screws (two).
Tighten the M2×10 pan-head machine screws (two).
Be aware that washers should be set to the M2×10 pan-head machine screws
(two).

SI/F interface connection

Two-wire serial interface (SI/F) for connecting printers and external display from UNIPULSE.
Connect from A11 and A12 of the external input/output connector.
The interface is nonpolarized and up to three external devices can be connected.
Wiring materials should use a two-core parallel cable, cabtyre cable (electric wire with covering
thickened for construction), etc. When a two-core parallel cable or a cabtyre cable is used, the
transmission distance is approximately 30m. When a two-core shielded twisted pair wire is used,
the transmission distance is approximately 300m.
Do not parallel it with AC lines and high-voltage lines. Doing so will cause malfunction.
INSTALLATION & CONNECTION
15
2TKPVGT
&KU
%QPXGTVGT
15

3 SETTING MODE CONFIGURATION

3 SETTING MODE
CONFIGURATION
Indicated value
display
HI/LO limit setting mode   (P21)
5:        A/D conversion speed select
P35
1: ch1 Digital filter (P37)
1: Motion Detect (P41)
3: Hysteresis (P45)
2: Near Zero (P47)
1: Comparison function
select (P48、P49)
ESC
F3(Mode3)
Ch Operation
F2(Mode2)
Calculation/common Operation
F1(Mode1)
Comparison/hold
FNC
ESC
HI/LO
H
I
Ch
1
Ch 2
OK
LO
PEAK
HOLD
H
I
LO
H
I
LO
H
I
Ch
1
Ch 2
OK
LO
PEAK
HOLD
H
I
LO
H
I
LO
H
I
Ch
1
Ch 2
OK
LO
PEAK
HOLD
H
I
LO
H
I
LO
2: SI/F function select (P62)
4:         D/Z synchronous mode
P39
3: Display frequency (P43)
7: Average sample number (P53)
6: Detection time (P57)
5: Calculation hold function
select (P50)
2: ch2 Analog filter
2: ch1 Analog filter (P36)
1: ch2 Digital filter
3: ch2 Zero Tracking
3: ch1 Zero Tracking (P40)

3-1. Setting mode composition

Chapter
3
SETTING MODE CONFIGURATION
16
16
3 SETTING MODE CONFIGURATION
F5(Mode5)
Option
BCO DAV/DAI232
2: Communication
condition (P72)
8:         Alarm LO limit
P33
7:         Alarm HI limit
P33
6:        Digital Zero limit
P32
5: ch2 Digital offset
4: ch2 Actual load
calibration
3:ch2 Equivalent input
calibration
2:ch2 Zero calibration
5: ch1 Digital offset P31
4: ch1 Actual load
calibra tion (P30)
3:ch1 Equivalent input
calibration (P28)
1:     Calibration function select
P23
F9(Mode9)
Protect/initialize
*As for an option setting, a setting item is different depending
F4(Mode4
Calibration setting
H
I
Ch1Ch 2
OK LO
PEAK
HOLD
HILO
HILO
Ch
1
HILO
H
I
Ch 2
OK LO
PEAK
HOLD
HILO
H
I
Ch1Ch 2
OK LO
PEAK
HOLD
HILO
HILO
2: ch1 Zero calibration P27
 Moreover, a set item is not displayed when there is no option.
2: Output rate (P68)
2: Zero scale value (P80 )
3: Delimiter (P72)
3: BCD / binary
select (P68)
3: Full scale value (P80 )
1: Communication
mode (P71)
1: Output data
select (P66)
1: Output data
select (P79)
4: Password (P97)
3: Initialize P97
2: Key protect (P96)
1: Setting protect (P95)
on the installed option.
Chapter
3
SETTING MODE CONFIGURATION
17
17
Loading...