Unipulse F130 User Manual

F130
LOADCELL INTERFACE
FOR OMRON SYSMAC CJ1
OPERATION MANUAL
06APR2009REV.1.00
WARNING
CAUTION
Thank you very much for purchasing our Loadcell Interface F130.
For good performance, and proper and safe use of the F130, be sure to read this operation manual
and properly understand the contents of it before use. Also, carefully keep this operation manual so
that it can be referred to at any time.
SAFETY PRECAUTIONS
For safety reasons, please read the following safety precautions thoroughly.
Please assign only persons with electrical technical knowledge to set up, service or inspect F130.
For the safe use of F130, safety precautions that should be complied with have been categorized
and indicated as " " and " " in this Operation Manual. The safety
precautions mentioned in this Operation Manual are crucial for safety. Therefore, use F130 only
after you have correctly understood the safety contents given.
WARNING
This sign forewarns the presence of hazards that could result in serious injury or fatality when incorrectly handled.
CAUTION
This sign forewarns the presence of hazards that could result in personnel injury or property damage when incorrectly handled.
I
WARNING
This sign forewarns the presence of hazards that could result in serious injury or fatality when incorrectly handled.
Warning on design
Warning on installation
Warning on wiring
Warning during startup and maintenance
Before using the F130 as described below, be sure to consult with our sales personnel.
- Use in environments not described in the operation manual.
- Use in environments that greatly affects human life and assets, such as medical devices,
transport equipment, recreational equipment, and safety devices.
Do not modify the F130. Doing so may cause fire or electric shocks.
Do not install in the following environments.
- Places containing corrosive gas or flammable gas.
- Where the product may be splashed with water, oil or chemicals.
Do not connect a commercial power source directly to the signal input/output terminals.
Be sure to ground the protective ground terminal.
Before performing the following, make sure that no power is applied.
- Wiring/connection of cables to the terminal boards.
For connection to the signal input/output terminals, check the signal names and pin
assignment numbers, and then carry out wiring properly.
Do not connect anything to unused terminal(s).
Before applying power, carefully check the wiring, etc.
Use at a proper power supply voltage.
Do not touch any terminal while applying power. Doing so may cause electric shocks or
malfunctions.
If the cover of the main body is opened, it may cause an electric shock internally. Even if the
power is off, the internal capacitor is charged. Contact us for internal inspection or repair.
In the case of smoke, an abnormal smell or strange sound, immediately turn off the power,
and disconnect the power cable.
II
CAUTION
This sign forewarns the presence of hazards that could result in personnel injury or property damage when incorrectly handled.
Caution on installation
Caution on wiring
Caution during startup and maintenance
Caution during transportation
Caution during disposal
Do not install in the following environments.
- Where the temperature/humidity exceeds the range of the specifications.
- Where the temperature changes severely or there is a danger of freezing or condensation.
- Places exposed to direct sunlight.
- Outdoors, or where the altitude exceeds 2000m.
- Dusty places
- Places containing large quantities of salt or iron powder.
- Where the main body is directly affected by vibrations or shocks.
Take adequate shielding measures when using at the following locations.
- Near a power line.
- Where a strong electric field or magnetic field is formed.
- Where static electricity, relay noise or the like is generated.
Install the F130 as far away as possible from devices that generate high frequency, high
voltage, large current, surge, etc. Also, carry out wiring separately from their power lines. Do not carry out parallel wiring and common wiring.
For sensors and external outputs use shielded cables.
For turning on/off the power, be sure to keep intervals of 5 seconds or more.
After power-on, be sure to warm up the F130 for 30 minutes or more before use.
If the F130 is not used by the specified method, its protective performance may be impaired.
Do not use it, broken down.
Maintenance
- When performing maintenance, disconnect the power.
- Do not wipe with a wet rag, or with benzine, thinner, alcohol, etc. Doing so may cause discoloration or deformation of the F130. In the case of heavy contamination, wipe off the contamination with a cloth after dipping it into a diluted neutral detergent and wringing it well, and then wipe with a soft, dry cloth.
When the F130 is shipped, spacers made of corrugated cardboard are used as cushioning
materials. Though it is factory-designed so that shocks can sufficiently be absorbed, breakage may result if shocks are applied when the spacers are reused for transportation. If you send the
F130 to us for repair, etc., take adequate measures against shocks by using polyurethane materials, etc., separately.
If you dispose of the product, handle it as industrial waste.
III
STATEMENT OF CONFORMATION TO EC DIRECTIVES
EN55016, EN61000-4-2, EN61000-4-3,
EN61000-4-4, EN61000-4-5, EN61000-4-6,
EN61000-4-8, EN61000-4-11
Unipulse F130 Loadcell Interface conforms to the EC Directives (based on Council of the European
Communities, and is allowed to affix CE mark on it.
* EMC Directives EN61131-2
The following notice must be observed when you install F130 unit.
1. F130 is an electrical device that are incorporated in machines and manufacturing installations.
F130 conforms to the related EMC standards so that the devices and machines into which F130
is built can more easily conform to EMC standards.
The actual F130 has been checked for conformity to EMC standards.
Whether these standards are satisfied for the actual system, however, must be checked by the
customer.
EMC-related performance will vary depending on the configuration, wiring, and other
conditions of the machine or device in which the F130 is installed.
The customer must, therefore, perform final checks to confirm that the overall machine or device
conforms to EMC standards.
2. Use shielded cables for connection with load cell.
RoHS-COMPLIANT PRODUCT
The parts and attachments (including the operation manual, packaging box, etc.) used for this unit
are compliant with the RoHS Directive restricting the use of hazardous substances with regard to
adverse effects on the environment and human body.
What is RoHS?
It is an abbreviation for Restriction on Hazardous Substances, which is implemented by the
European Union (EU). The Directive restricts the use of six specific substances in electric and
electronic equipment handled within EU borders. The six substances are lead, mercury, cadmium,
hexavalent chromium, PBB (polybrominated biphenyls), and PBDE (polybrominated diphenyl
ethers).
IV
CONTENTS
1APPEARANCE DESCRIPTION ..............................................1
1-1Front Panel ................................................................................................. 1
1-2Status LED ................................................................................................. 1
1-3Unit Number Select Switch ....................................................................... 2
2CONNECTION METHOD ........................................................3
2-1F130 Terminal Block Layout ..................................................................... 3
2-2Sensor Connection .................................................................................... 3
2-2-1 4-wire Connection ....................................................................................... 3
2-2-2 6-wire Connection ....................................................................................... 4
2-2-3 Sensor Cable Coloration ............................................................................ 4
CONTENTS
2-3Connections to Terminal Block ................................................................ 5
3DATA EXCHANGE WITH CPU ..............................................6
3-1High Performance I/O Unit Restart Flag .................................................. 7
3-2Relay Area .................................................................................................. 7
3-2-1 Allocation of MeasurementHold Data .................................................... 7
3-2-2 OUT (CPU Unit F130) .............................................................................. 8
3-2-3 IN (F130 CPU Unit) .................................................................................. 9
3-2-4 Register Allocations for Initial Settings .................................................. 11
4CALIBRATION METHOD .....................................................12
4-1What is Calibration? ................................................................................ 12
4-1-1 About Calibration Methods ...................................................................... 12
4-2Actual Load Calibration and Equivalent Input Calibration Procedure 13
4-3Preparation for Calibration ..................................................................... 14
4-3-1 Setting Initial Data ..................................................................................... 14
4-3-2 Calibration Mode Selection ...................................................................... 15
4-4Zero Calibration ....................................................................................... 15
4-5Actual Load Calibration .......................................................................... 15
4-6Equivalent Input Calibration ................................................................... 16
4-7Calibration Error ...................................................................................... 16
V
CONTENTS
5MEASUREMENT SETTINGS ...............................................17
5-1Analog Filter Selection (m+2) ................................................................. 17
5-2Digital Filter Count Setting (m+3) ........................................................... 17
5-3Digital Offset (m+4) ................................................................................. 17
6MEASUREMENTHOLD FUNCTIONS ...............................18
6-1Measurement Function ........................................................................... 18
6-2Hold Functions ........................................................................................ 18
6-2-1 Peak Hold ................................................................................................... 18
6-2-2 Bottom Hold ............................................................................................... 19
6-2-3 Average Hold ............................................................................................. 19
6-2-4 Application of the Hold Clear Function ................................................... 20
7BLOCK DIAGRAM ................................................................21
8DIMENSIONS ........................................................................22
9INSTALLATION ....................................................................23
9-1Connection with CJ1 Unit ....................................................................... 23
9-2DIN Rail Installation ................................................................................. 24
10CAUTION ABOUT POWER CONSUMPTION
AND SYSTEM CONFIGURATION ......................................26
11SPECIFICATIONS ..............................................................27
11-1Analog Section ...................................................................................... 27
11-2Display .................................................................................................... 27
11-3Setting .................................................................................................... 28
11-4Terminal Block Layout .......................................................................... 28
VI
11-5General Specifications .......................................................................... 28

1. APPEARANCE DESCRIPTION

Status LED
Terminal block
Unit number select switch
B
1A
1
M
A
C
H
N
o
.
1
0
1
1
0
0
F130
R
U
N
ER
C
E
R
H

1-1. Front Panel

1.APPEARANCE DESCRIPTION

1-2. Status LED

CAUTION
Terminal block can be detached by pulling down the lever. Normally, the lever should be in upper position.
Always turn off the CJ1’s voltage source before trying to attach/detach the terminal block.
Status LEDs display the current operating status of the unit. The following table summaries their
names and meanings.
LED Name State Description
RUN
(Green)
ERC
(Red)
ERH
(Red)
Operation status
Error detected by the unit
CPU error
LED ON Normal operation
LED OFF Data exchange with CPU has been aborted
LED ON Error occurred in F130
Error (calibration error / setting error) or overflow
LED Blink
LED OFF Normal operation
LED ON Error occurred during data exchange with CPU
LED OFF Normal operation
(A/D input range over / measurement value over) occurred in F130.
1
1.APPEARANCE DESCRIPTION

1-3. Unit Number Select Switch

F130 acts as a Special I/O unit for Omron SYSMAC CJ1 series PLC. The data exchange between
CPU unit and F130 are carried out by the high performance relay area of I/O and DM area.
Relay and DM number used by F130 can be selected by the unit number selection switch located on
the front panel.
Switch
No.
0 0 2000 to 2009CH D20000 to 20099
1 1 2010 to 2019CH D20100 to 20199
2 2 2020 to 2029CH D20200 to 20299
3 3 2030 to 2039CH D20300 to 20399
4 4 2040 to 2049CH D20400 to 20499
5 5 2050 to 2059CH D20500 to 20599
6 6 2060 to 2069CH D20600 to 20699
7 7 2070 to 2079CH D20700 to 20799
8 8 2080 to 2089CH D20800 to 20899
9 9 2090 to 2099CH D20900 to 20999
10 10 2100 to 2109CH D21000 to 21099
n n
95 95 2950 to 2959CH D29500 to 29599
Unit No.
Channel number allocated
to the I/O unit relay area
2000+n × 10 to
2000+n × 10+9CH
DM number allocated
to the I/O unit DM area
D20000+n × 100 to
D20000+n × 100+99
Unit number must be unique: duplicate use of the same number for multiple of
I/O units will cause operation error “Error: Duplicate use of unit No.” (“Duplicate use of unit No.”
in programming console), disabling normal operation (A40113 turns ON).
2

2. CONNECTION METHOD

SHIELD B1
+SIG B2
-SIG B3
+EXC B4
-EXC B5
RSV B6
RSV B7
RSV B8
RSV B9
A1 RSV
A2 RSV
A3 RSV
A4 RSV
A5 RSV
A6 RSV
A7 RSV
A8 RSV
A9 RSV
+IN
-OUT
-IN
+EXC
-EXC
+SIG
-SIG
SHIELD
+OUT
Red
Black
Green
White
B4
B5
B2
B3
B1
Color coding used in UNIPULSE supplied 4-core cable

2-1. F130 Terminal Block Layout

2.CONNECTION METHOD

2-2. Sensor Connection

2-2-1. 4-wire Connection

F130 input uses an accurate 4-wire sensor connection. Use shielded 4-core cable for this connection
and route it apart from noisy lines (power and digital devices) and AC power lines.
3
2.CONNECTION METHOD
Red
Yellow
Black
Orange
Green
White
+IN
-OUT
-IN
+EXC
-EXC
+SIG
-SIG
SHIELD
+OUT
B4
B5
B2
B3
B1
Color coding used in UNIPULSE supplied 6-core cable

2-2-2. 6-wire Connection

For connecting a 6-wire strain gauge type sensor, connect +EXC and +S, and -EXC and -S,
respectively, on the terminal block as shown below.
CAUTION
The strain gauge type sensor excitation voltage of the F130 is 2.5V or 10V (selectable by setting). Set within the maximum excitation voltage of the strain gauge type sensor you use, and then make connection. Use exceeding the maximum excitation voltage may cause abnormal heating and damage.

2-2-3. Sensor Cable Coloration

Sensor cable coloration differs with manufacturers.
See the sensor operation manual (or test report), and check the signal names and coloration for
proper connection.
4

2-3. Connections to Terminal Block

Within 6mm Within 6mm
Pressure terminal
M3 screw
Use pressure terminals to connect cables to the terminal. Tighten the pressure terminal securely to
the terminal block using a M3 screw. Correct tightening torque is 0.5Nm.
The width of the pressure terminal should be 6mm or less (see the figure below).
2.CONNECTION METHOD
CAUTION
A1 to A9 and B6 to B9 are unnecessary in this specification, please do not connect input and output signal lines to them. Do not apply the voltage from the outside, or do not short-circuit with parts. It may cause malfunction of F130 and external equipment.
5

3.DATA EXCHANGE WITH CPU

CJ series CPU unit F130
Measurement
Hold data
2000+n
×
10
2000+n
×
10+9
High performance I/O unit relay area
10CH
D20000+n
×
100
D20000+n
×
100+99
Data memory (DM) area
n: unit number
Measurement value,
(Peak, Bottom, Average)
Initial setting data
Measurement data, Hold data
Initial setting data
I/O refresh
Power-on
or
etc.
(Excitation voltage,
Calibration data,
Filter setting etc.)
unit restart
Initial setting
100CH
data
data
Hold value
3. DATA EXCHANGE WITH CPU
The F130 exchanges data with CPU using the following devices: High performance I/O unit relay
area (measurement value and hold value) and high performance I/O unit DM area (Initial setting
data).
Measurement valueHold value
Analog signal from the strain gauge type sensor is converted into digital data. Measurement・Hold data is allocated to one of the relay area of the CPU units’ high performance I/
O unit based upon the unit number. The data is exchanged regularly every time I/O refresh takes
place.
Initial setting data
Parameter settings for using F130 as a
A set of initial setting data for each F130 is allocated, based on its unit number, to a DM area for
high performance I/O unit (inside CPU unit), and is written to this area when the unit is powered
on, or a refresh cycle is started.
Loadcell Interface
.
6

3-1. High Performance I/O Unit Restart Flag

When the user restart the unit after he has modified data memory or removed the cause of failure,
the user have to either power up the CJ1 main unit again, or changing the high performance I/O unit
restart flag in the following sequence: OFF ON OFF.
High performance I/O unit restart flag
Relay number Function
A50200 0th unit restart flag
A50201 1th unit restart flag
3.DATA EXCHANGE WITH CPU
A50215 15th unit restart flag
A50300 16th unit restart flag
A50715 95th unit restart flag

3-2. Relay Area

3-2-1. Allocation of MeasurementHold Data

OUT (CPU unit F130)
1514131211109876543210
Factory-
Mode
adjusted
selection
n CH
request
operation
request
IN (F130 CPU unit)
1514131211109876543210
Factory-
Mode selection response
adjusted
operation
response
n+1 CH
n+2 CH Measurement value (-32768 to +32767)
Setting
error
occurre-
nce
occurre-
Cal.
error
nce
Equiv.
input
cal.
request
Equiv.
input
cal.
response
Actual
load
cal.
request
Actual
load cal.
in
progress
Zero
cal.
request
Zero
cal. in
progress
Survival
conf.
request
Survival
conf. response ((n) 07 is inverted.)
Normally
ON
(Normally
1)
Restart each unit by switching OFF ON OFF
Hold value clear
request
Measure
overflow
ment value
A/D input range
over
Hold value clear
response
Hold
request
(section
spec.)
Hold
response
Digital
zero
request
Digital
zero
response
n+3 CH
n+4 CH Peak hold value (-32768 to +32767) / In calibration mode: strain gauge type sensor rated output value
n+5 CH Bottom hold value (-32768 to +32767) / In calibration mode: strain gauge type sensor rated capacity value
n+6 CH Average value (-32768 to +32767) / In calibration mode: strain gauge type sensor output value at the Zero calibration point
n+7 CH Average count (0,1 to 30000) / In calibration mode: calibration error number
n+8 CH Unused
n+9 CH Unused
Data update counter (0 to 65535)
n = 2000 + unit number × 10   * (n) 08 to 10 are valid only in the calibration mode.
7
3.DATA EXCHANGE WITH CPU

3-2-2. OUT (CPU Unit F130)

Digital zero request
ON edge (0 1) zeros the indicated value.
Hold request (section specification)
While this bit is ON (1), each hold value is detected and updated according to hold conditions.
Hold value clear request
ON edge (0 1) clears the hold value. The output value becomes the indicated value of the signal-
input time.
Survival confirmation request
By changing this bit, the data transfer state between the CPU unit and F130 can be confirmed. A response is made to the 7th bit of n+1 CH as this bit is inverted (0 ⇔ 1).
Zero calibration request
ON edge (0 1) initiates Zero calibration.
However, this is valid only when the calibration mode is selected by Mode selection.
Actual load calibration request
ON edge (0 1) initiates Actual load calibration.
However, this is valid only when the calibration mode is selected by Mode selection.
Equivalent input calibration request
ON edge (0 1) initiates Equivalent input calibration.
However, this is valid only when the calibration mode is selected by Mode selection.
Factory-adjusted operation request
This bit is for factory-adjusted operation. Be sure to keep this OFF. If this bit is changed (OFF ON), normal operation may not be performed, or data may not be
recovered in some cases.
Mode selection
Select the operation mode of the F130.
Mode selection Operation mode
OFF Normal operation mode
ON Calibration mode
8
3.DATA EXCHANGE WITH CPU

3-2-3. IN (F130 CPU Unit)

Digital zero response
Returns the state of the digital zero request bit.
Hold response
Returns the state of the hold request bit.
Hold value clear response
Returns the state of the hold value clear request bit.
A/D input range over
This bit becomes "1" when the input from the strain gauge type sensor is out of the input range of
the A/D converter.
Measurement value overflow
This bit becomes "1" when the indicated value is smaller than -32768, or larger than +32767.
Normally ON
Always returns "1."
Survival confirmation response
Returns the state of the survival confirmation request (n) 07 as it is inverted.
Zero calibration in progress
This bit becomes "1" while Zero calibration is in progress.
Actual load calibration in progress
This bit becomes "1" while Actual load calibration is in progress.
Equivalent input calibration response
Returns the state of the Equivalent input calibration request bit.
Calibration error occurrence
This bit becomes "1" if Zero calibration, Actual load calibration, or Equivalent input calibration
does not end properly with a calibration error.
Setting error occurrence
This bit becomes "1" when the initial set value assigned in the data memory area is out of the setting range.
Factory-adjusted operation response
Returns the state of the factory-adjusted operation request bit.
9
3.DATA EXCHANGE WITH CPU
Mode selection response
Returns the state of the mode selection request bit.
Set value (n+2)
Returns the indicated value of A/D-converted measurement.
-32768 to +32767; two's-complement number.
Data update counter (n+3)
Returns a numerical value incremented by one each time the indicated value is updated (3000
times/sec.).
0 to 65535; straight binary.
Peak hold value / strain gauge type sensor rated output value (n+4)
In normal operation mode
Returns the maximum measurement value in the section specified by the hold request bit.
When the hold request bit turns OFF, data update is stopped, and the maximum measurement
value of that time is held.
By the hold value clear request bit, the maximum measurement value is cleared, and the real-
time measurement value is returned.
-32768 to +32767; two's-complement number
In calibration mode
When the calibration mode is established by the mode selection bit, the initially set value "strain
gauge type sensor rated output value" of the time when the calibration was executed by each
calibration request bit is read and returned.
However, when Span calibration is executed, the input value from the strain gauge type sensor
is returned.
* After power-on, "0" is returned until the calibration is executed.
Bottom hold value / strain gauge type sensor rated capacity value (n+5)
In normal operation mode
Returns the minimum measurement value in the section specified by the hold request bit.
When the hold request bit turns OFF, data update is stopped, and the minimum measurement
value of that time is held.
By the hold value clear request bit, the minimum measurement value is cleared, and the real-
time measurement value is returned.
-32768 to +32767; two's-complement number
In calibration mode
When the calibration mode is established by the mode selection bit, the initially set value "strain
gauge type sensor rated capacity value" of the time when the calibration was executed by each
calibration request bit is read and returned.
* After power-on, "0" is returned until the calibration is executed.
10
3.DATA EXCHANGE WITH CPU
Average value / strain gauge type sensor output value at the Zero calibration point (n+6)
In normal operation mode
Returns the average measurement value in the section specified by the hold request bit.
When the hold request bit turns OFF, data update is stopped, and the average measurement
value of that time is held.
By the hold value clear request bit, the average measurement value is cleared, and the real-time
measurement value is returned.
-32768 to +32767; two's-complement number
In calibration mode
When the calibration mode is established by the mode selection bit, the strain gauge type sensor
input value of the time when the calibration was executed by the Zero calibration request bit is
returned.
* After power-on, "0" is returned until the Zero calibration is executed.
Average count / calibration error number (n+7)
In normal operation mode
Returns the A/D conversion count in the section specified by the hold request bit as the number
of data used for average calculation.
By the hold value clear request bit, the average count value is cleared, and "0" is returned.
1 to 32767 ("0" when cleared); two's-complement number
In calibration mode
When the calibration mode is established by the mode selection bit, the calibration error number
of the time when the calibration was executed by each calibration request bit is returned.
* After power-on, "0" is returned until the calibration is executed.

3-2-4. Register Allocations for Initial Settings

1514131211109876543210
m
m+1
m+2
m+3
m+4
m+5
m+6
m+7
m+8
m+9
Sensor rated capacity value (1 to 32000) * Used for Actual load calibration or Equivalent input calibration.
* m+10 ~ m+99: Unused
Analog filter selection (0:3Hz, 1:10Hz, 2:30Hz, 3:100Hz, 4:300Hz, 5:1kHz)
Digital filter count setting (1 to 512) * Digital filter: Moving-average
Sensor rated output value (50 to 3100) * Used for 0.050 to 3.100mV/V Equivalent input calibration.
Excitation voltage selection (0:2.5V, 1:10V)
Minimum scale division (1 to 100)
Digital offset (-30000 to +30000)
Unused
Unused
Unused
11

4.CALIBRATION METHOD

Just connecting a strain gauge type sensor to F130 …
F130 indicates random values
After calibration ….
The F130 and strain gauge type sensor
function as a weighing device
10kN
10kN
A data sheet is attached to a strain gauge type sensor at the time of purchase. For Equivalent input calibration, the "rated capacity value" and "rated output value" described on the data sheet are set to the F130: therefore, it should be carefully stored after calibration.
Attention
4. CALIBRATION METHOD

4-1. What is Calibration?

Calibration is performed for matching the F130 to a strain gauge type sensor.
That is, F130 must be properly adjusted to indicate 10.00kN when an actual load of 10kN is placed
on the strain gauge type sensor.

4-1-1. About Calibration Methods

12
There are two methods of Span calibration: Actual load calibration and Equivalent input
calibration.
Actual load calibration
By this method, calibration value is registered by the value of the actual load applied to the strain
gauge type sensor.
Error-less accurate calibration can be performed.
Equivalent input calibration
By this method, calibration value is registered from the rated output value (mV/V) and rated capacity
value of the strain gauge type sensor. This calibration method using no load can simplify work.
4.CALIBRATION METHOD
Calibration mode release
Span Calibration (Actual load calibration/
Calibration mode selection
F130 Power on, or restart
Input initial setting data
Zero Calibration
Equivalent input calibration)
Zero Calibration

4-2. Actual Load Calibration and Equivalent Input Calibration Procedure

Steps required to perform Actual load calibration and Equivalent input calibration are as follows:
Write initial setting data to the DM area that corresponds to
MACH No.
Register initial setting data to F130. To do this, either power on
the F130, or restart it by toggling restart flag in the following sequence: OFF → ON → OFF.
Set the mode selection bit (15th bit of nCH) to "1" to establish
the calibration mode.
Register initial zero value by setting the Zero calibration bit
(8th bit of n CH) to “1”.
- Actual load calibration
Register a span by applying actual load to the strain gauge type
sensor.
- Equivalent input calibration
Execute Equivalent input calibration.
(Set the Equivalent input calibration bit (10th bit of nCH) to "1.")
Perform the Zero calibration procedure again if required.
Set the mode selection bit (15th bit of nCH) to "0" to restore the
normal operation mode.
13
4.CALIBRATION METHOD
Since calibration value data are stored separately for excitation voltage 10V and
2.5V, it should always be determined before performing calibration operation, and it should not be changed after calibration. If the maximum excitation voltage of the strain gauge type sensor is under 10V, be sure to select 2.5V.
Attention
Set each set data properly within the setting range. In the case of setting out of the setting range, normal operation may not be performed with measurement value and hold value.
* If out of the setting range, "1" is returned to the 12th bit of (n+1)CH "setting error
occurrence."
Attention

4-3. Preparation for Calibration

4-3-1. Setting Initial Data

Minimum data items required for performing proper calibration are: Excitation voltage, Minimum
scale division, Sensor rated output value, and Sensor rated capacity value. Initial setting data are
registered when F130 is powered on or it restarts.
Excitation voltage selection (m)
Select the bridge voltage supplied to the strain gauge type sensor.
The bridge voltage can be selected from 2.5V and 10V.
After this setting, be sure to perform calibration.
(Setting range 0: 2.5V, 1: 10V)
Minimum scale division (m+1)
Set the increment of the indicated value.
(Setting range 1 to 100)
Sensor rated output value (m+5)
Set the rated output value of the strain gauge type sensor for Equivalent input calibration.
Since the set value is reflected when the calibration is executed (ON edge of the Equivalent input
calibration request), set before execution.
(Setting range 50 to 3100 (0.050 to 3.100mV/V))
Sensor rated capacity value (m+6)
Set the indicated value corresponding to the actual load for Actual load calibration.
Set the rated output value of the strain gauge type sensor for Equivalent input calibration.
(Setting range 1 to 32000)
14

4-3-2. Calibration Mode Selection

15 14 13 12 4 3 2 1 0
n CH
Mode
selection
0: Normal operation mode 1: Calibration mode
On the F130, Zero calibration and Span calibration can be operated by selecting the calibration
mode with the mode selection bit. Use the normal operation mode except during calibration.

4-4. Zero Calibration

Follow the steps below to define correct initial value:
1. Make sure that the strain gauge type sensor is free from unnecessary load, such as foreign object
placed on the weighing plate, or contact with peripheral devices.
2. Make sure that the measurement value (n+2 CH) is stable without fluctuations.
3. Toggle the Zero calibration bit (8th bit of n CH) from 0 → 1.
4.CALIBRATION METHOD
4. F130 starts Zero calibration procedures when it acknowledges an ON edge (0 1) of the Zero
calibration bit.
5. Zero calibration in progress bit (8th bit of n+1 CH) turns on “1”. Do not touch the sensing
portion (sensor) while this bit in on.
6. Make sure measurement value (n+2 CH) shows “0”.
7. Toggle the Zero calibration bit (8th bit of n CH) back to “0”.

4-5. Actual Load Calibration

Apply an actual load to the strain gauge type sensor and register the span.
1. Apply an actual load to the strain gauge type sensor.
2. Make sure that the strain gauge type sensor is free from unnecessary load, such as foreign object
placed on the weighing plate, or contact with peripheral devices.
3. Make sure that the measurement value (n+2 CH) is stable without fluctuations.
4. Toggle the Actual load calibration bit (9th bit of n CH) from 0 → 1.
5. F130 starts Actual load calibration procedures when it acknowledges an ON edge (0 1) of the
Actual load calibration bit.
6. Actual load calibration in progress bit (9th bit of n+1 CH) turns on “1”. Do not touch the sensing
portion (sensor) while this bit in on.
7. Make sure that the measurement value (n+2 CH) exactly coincides with the sensor rated capacity
value (m+6).
8. Toggle the Actual load calibration bit (9th bit of n CH) back to “0”.
15
4.CALIBRATION METHOD

4-6. Equivalent Input Calibration

1. Toggle the Equivalent input calibration bit (10th bit of n CH) from 0 → 1.
2. F130 starts Equivalent input calibration procedures when it acknowledges an ON edge (0 → 1)
of the Equivalent input calibration bit.
3. Toggle the Equivalent input calibration bit (10th bit of n CH) back to “0”.

4-7. Calibration Error

If calibration operation is performed without properly making strain gauge type sensor input or
initial setting to the F130, a calibration error occurs. Check the calibration error number (n+7 CH),
remove the cause, and then perform calibration operation again.
When the calibration error occurs, "1" is returned to the 11th bit of n+1 CH "setting error
occurrence."
Calibration Error 2
This indicates a state in which Zero calibration cannot be performed because an electric signal
exceeding the input voltage range is input on the positive side. Check the cable for breaks or wrong
wiring.
Calibration Error 3
This indicates a state in which Zero calibration cannot be performed because an electric signal
exceeding the input voltage range is input on the negative side. Check the cable for breaks or wrong
wiring.
Calibration Error 5
Sensor rated capacity value is preset to “0”. Select a proper non-zero value.
Calibration Error 6
The output of the strain gauge type sensor is too small to perform calibration. Check to see if load is
applied properly, or the strain gauge type sensor has such a capability that can make appropriate
output, and then perform Span calibration again.
16
Calibration Error 8
Strain gauge type sensor output is outside the F130’s allowable sensor input range. Check if the
sensor is properly loaded, and the its rated output falls within the F130’s sensor input range.
Perform Span calibration again. Then, perform Span calibration again.

5. MEASUREMENT SETTINGS

Set each set data properly within the setting range. In the case of setting out of the setting range, normal operation may not be performed with measurement value and hold value. * If out of the setting range, "1" is returned to the 12th bit of n+1 CH "setting error
occurrence."
Attention

5-1. Analog Filter Selection (m+2)

A low-pass filter is provided for filtering input signals from the strain gauge type sensor and
cancelingnoise components.
The cut-off frequency can be selected from 3Hz, 10Hz, 30Hz, 100Hz, 300Hz, 1kHz.
With an increase in the cut-off frequency, the response becomes faster, but noise components may
be indicated.
(Setting range 0: 3Hz, 1: 10Hz. 2: 30Hz, 3: 100Hz, 4: 300Hz, 5: 1kHz)

5-2. Digital Filter Count Setting (m+3)

This filter calculates moving average of A/D converter output to reduce indicated value fluctuation.
Set by the sampling count of the past data to perform moving average process.
Larger averaging width will enhance indicated value stability, but reduce response performance.
Vice versa, shorter averaging width reduces indicated value stability, but enhance response velocity.
(Setting range 1 to 512)
5.MEASUREMENT SETTINGS

5-3. Digital Offset (m+4)

This function always subtracts the digital offset set value from the indicated value.
This function is convenient when zero cannot be obtained with no load for some reason or for
offsetting. When not using, set "00000".
(Setting range -30000 to +30000)
17

6.MEASUREMENTHOLD FUNCTIONS

Hold request
Hold value
Peak
Measurement
Detection
OFF
ON OFF
ON
t1 t2
t3
value
clear request
6. MEASUREMENTHOLD FUNCTIONS

6-1. Measurement Function

Measurement value and data update counter
The F130 performs A/D conversion at a rate of 3000 times/sec to calculate the measurement value.
Each time the measurement value is updated, the data update counter is incremented by one. Since
the measurement value and counter value are returned in pairs, how often you see the measurement
data of the F130 can be known from the progression of the counter value.
Measurement
value
9999 3245
10001 3248 3
10000 3251 3
9998 65535
10000 3 4
9999 6 3

6-2. Hold Functions

The F130 has three hold functions: peak hold, bottom hold, and average hold, and these hold
functions operate simultaneously.

6-2-1. Peak Hold

The maximum value (peak) in the positive direction of the specified section is held.
Data update
counter
Counter
difference
According to the difference from the previous
counter value, you can see that the rate is once out of three times (1000 times/sec.).
Goes back to "0" after counting up to 65535.
18
t1: Delay time between the instant when the hold request turns ON and the instant when the mea-
surement value hold detection is started
t2: Delay time between the instant when the hold request turns OFF and the instant when the mea-
surement value hold detection is stopped
t3: Delay time between the instant when the hold clear request turns OFF and the instant when the
hold value is released
6.MEASUREMENTHOLD FUNCTIONS
Hold request
Bottom
Detection
OFF
ON OFF
ON
t1 t2
t3
Measurement value
Hold value clear request
Hold request
Average
Detection
OFF
ON OFF
ON
t1 t2
t3
Measurement value
Hold value clear request

6-2-2. Bottom Hold

The maximum value (bottom) in the negative direction of the specified section is held.
t1: Delay time between the instant when the hold request turns ON and the instant when the mea-
surement value hold detection is started
t2: Delay time between the instant when the hold request turns OFF and the instant when the mea-
surement value hold detection is stopped
t3: Delay time between the instant when the hold clear request turns OFF and the instant when the
hold value is released

6-2-3. Average Hold

The average value of the specified section is held.
t1: Delay time between the instant when the hold request turns ON and the instant when the mea-
t2: Delay time between the instant when the hold request turns OFF and the instant when the mea-
t3: Delay time between the instant when the hold clear request turns OFF and the instant when the
surement value hold detection is started
surement value hold detection is stopped
hold value is released
19
6.MEASUREMENTHOLD FUNCTIONS
< About average value measurement time >
What can be specified as a section for average value hold detection is up to 10 seconds (average count: 30000). Even if the hold request bit is ON exceeding that time, the average value in 10 seconds is held.
Point
< About the request bit delay times (t1 to t3) >
Since the delay times change depending on the system environment, sufficient evaluation should be performed before use in response to the actual status of use.
Attention
Hold request
Peak
Detection
OFF
ON OFF
ON
Peak
Measurement value
Hold value clear request

6-2-4. Application of the Hold Clear Function

If the hold clear is turned ON when the hold request is ON (in the detection section), the previous
hold value is once cleared, and then hold detection proceeds.
20

7. BLOCK DIAGRAM

strain gauge
NOV.RAM
512 byte
LPF
Reference
voltage
A/D
24 bit
Flash-ROM
256k byte
RAM
16k byte
Preamplifier
fc selection Excitation voltage selection
Isolator
PORT
SCI
PORT
PORT
BUS
16bit CPU
* For calibration data save
Electrically Isolated
DC-DC Converter
+5V
+12V
-12V
Logic (CPU, ASIC)
Analog
4-wire
CS BUS
(Vref)
Sensor voltage
ASIC only for High performance
I/O unit
type sensor
At DC10V (30mA max)
single 350Ωsensor
At DC2.5V (30mA max)
single 120Ωsensor or
four 350Ωsensors in parallel
7.BLOCK DIAGRAM
21

8.DIMENSIONS

˴ ˴
˴
˴
Unit: mm
8. DIMENSIONS
22

9. INSTALLATION

P
A
205R
P
O
W
E
R
INPUT
A
C1
00-24
0V
L2/N
L1
DC24
V
A
C2
4
0V
OUTP
U
T
RUN
P
E
RIP
H
E
R
A
L
ERR/
AL
M
RUN
I
NH
COM
M
PRPHL
C ON
T
R
OLL
E R
C
J1G-C
P
U
4
4
S
Y
S
MA
C
P
R
O
G
R
A
M
MA
B
L
E
PORT
O
P
E
N
B
U
S
Y
M
C
P
WR
Connector
Fooking hole
Fook
PA205
R
POW
E
R
IN
PU
T
A
C
1
0
0-2
4
0
V
L
2
/
N
L
1
D
C
24
V
A
C
2
4
0V
OUTP
U
T
R
U
N
PERIPHERAL
E
RR/
A
L
M
RUN
I
NH
COMM
PRPHL
C
O
NT
R
O
L
L
E R
C
J1G
-
C
P
U
44
S
YS
M
A
C
P
ROG
R
A
M
M
A
B
L
E
P
O
R
T
O
P E
N
B
U S
Y
MC
P
W
R
Slide the slider towards the back of the unit until it “clicks”.
Locked
Released
Slider

9-1. Connection with CJ1 Unit

Multiple of CJ1 units can be connected by simply engaging connectors and locking the slider. Add
an end cover to the unit installed to the right most unit.
1. Engage the connector properly to connect units.
9.INSTALLATION
CJ series do not require base unit. Each unit can be connected to each other using the integrated
side-mount connector.
CAUTION
Never fail to turn off the system before starting system assembly.
When replacing a unit, remove the assembled unit as a whole before replacing one
of them.
2. Slide the two yellow sliders on the upper and lower end of the unit until they “clicks” to lock securely.
Lock the unit securely by moving the slider to its full stroke. Insufficient engagement can hinder the system from proper operation.
CAUTION
23
9.INSTALLATION
End cover
DIN rail pin
Released” position
DIN
1
2
DIN rail
3. Add an end cover to the right most unit.
You must install an end cover to the right most units. Without it, CJ1’s CPU unit issues “I/O bus error” (Operation Halt Error) and will not start operation (neithre “operation” nor “monitoring” mode).

9-2. DIN Rail Installation

CAUTION
CJ series can only be mounted to DIN type rail. It can no be installed using screws.
1. Set the DIN rail pins on the rear of the unit to “Released” position.
2. Hook the CJ unit’s upper protrusions on the DIN rail ( ), and engage the entire notch to the rail ( ).
24
3. Lock all the DIN rail pins by pushing upwards.
DIN
DIN rail pin
1
2
End plate
For details on the handling of CJ series units, see “OMRON CJ Series: CJ1/CJ1-H/ CJ1M CPU Unit User’s Manual (Setup)” provided by OMRON.
Attention
4. Secure the unit by installing a end plate.
9.INSTALLATION
To fix CJ1 assembly securely, sandwitch the assembly using a pair of endplates from both sides.
Fook lower notch of the end plate to the lower plate of the rail ( ), and then engage the upper notch, then pull the end plate to lower direction ( ). Tighten the
screw to fix them serurely.
25

10.CAUTION ABOUT POWER CONSUMPTION AND SYSTEM CONFIGURATION

For the system configuration of the CJ series and the power consumption of each unit, see "OMRON CJ series: CJ1/CJ1-H/CJ1M CPU Unit User's Manual (Setup)," etc.
Attention
10. CAUTION ABOUT POWER CONSUMPTION AND SYSTEM CONFIGURATION
Design your system with sufficient consideration given to power consumption so that the total
power consumption of the F130 and units of the CJ series (CPU unit, Configuration units, etc.) does
not exceed the maximum supply capability of the power supply unit.
F130 power consumption: 0.46 [A]
The number of connectable units is up to 10 units per CPU rack
the CJ series. (It can be expanded to 40 units depending on the CPU unit.)
* The CJ-series CPU rack consists of a CPU unit, power supply unit, configuration units (Basic I/O
units, Special I/O units(F130 included), and CPU bus units), and an end cover.
*
according to the specifications of
26

11. SPECIFICATIONS

11-1. Analog Section

Number of sensor connection ch 1ch
Sensor connection terminal Detachable terminal block
Sensor connection method 4-wire input (+SIG, -SIG, +EXC, -EXC, SHIELD)
Sensor excitation voltage DC10V ± 5%, source current max. 30mA
Signal input range -3.0 to +3.0mV/V
11.SPECIFICATIONS
(single 350Ωsensor)
or DC2.5V ± 5%, source current max. 30mA (single 120Ωsensor or four 350Ωsensors in parallel)
Auuracy Non-linearity: Within 0.02%FS (at a 3.0mV/V input)
Analog filter Low-pass filter (-6dB/oct)
A/D converter Rate: 3000 times/sec.

11-2. Display

Status LED RUN (Green): Operation status
Zero drift: Within 0.5μV/ ℃ RTI Gain drift: Within 25ppm/
Cut off frequency: Selectable from 3, 10, 30, 100, 300, 1k Hz
Resolution: 24bit binary
(Effective resolution: approx. 16 to 18 bits)
* When the effective resolution is 16 bits,
the solution results in counting approx. 30000
with respect to 3.0mV/V.
However, the bit stability depends on the
averaging filter setting.
(Turns on when Normal operation)
ERC (Red): Error detected by the unit
(Turns on
blinks when an error occurs in the F130.)
ERH (Red): CPU error
(Turns on when an alarm occurs in data
exchange with the CPU unit.)
when an alarm occurs or
27
11.SPECIFICATIONS
SHIELD B1
+SIG B2
-SIG B3
+EXC B4
-EXC B5
RSV B6
RSV B7
RSV B8
RSV B9
A1 RSV
A2 RSV
A3 RSV
A4 RSV
A5 RSV
A6 RSV
A7 RSV
A8 RSV
A9 RSV

11-3. Setting

High performance I/O unit No.
Calibration valueset value setting method
Calibration value Stored in NOV RAM (non-volatile RAM)
Set value Measurement starts with the set values (DM area) written from the

11-4. Terminal Block Layout

Detachable terminal block (M3 screw)
Decimal 2-digit rotary switch (unit No. 0 to 95)
Made from the CPU unit via the data memory (DM) area
defined for the F130.
CPU unit after power-on. (Not stored on the F130 side)
28
* Never connect anything to the RSV terminals.

11-5. General Specifications

Internal consumption current
DC5V, Approx. 0.46A (At excitation voltage DC10V, single 350Ω sensor)
Operating conditions Operation temperature: 0 to +55
Accuracy compensation applicable temperature:
Dimensions 31(W) × 90(H) × 65(D) mm (protrusions excluded)
Weight Approx. 160g
Storage temperature: -20 to +75
Humidity: 90%RH or less (non-condensing)
0 to +40
Loading...