Rights Reserved. No reproduction without written authorization.
Operator’s Manual
Version 3.2
DATE: March 2019
Page 3
table of contents
Laser Safety 1
Warranty information vii
Sales and Applications viii
Customer Service viii
Technical Support viii
Reference materials viii
Hazard Information 11
Terms 11
General hazards 12
Other hazards 14
Disposal 14
Additional laser safety information 14
Agency compliance 18
Center for Devices and Radiological Health (CDRH) requirements 18
OEM model 18
Federal Communications Commission (FCC) 19
Requirements 19
European Union (EU) requirements 20
Introduction 26
Firestar nomenclature 26
Unpacking 26
Getting Started 2
Introduction 26
Firestar nomenclature 26
Unpacking 27
Incoming inspection 27
Packaging guidelines 27
Removing the lifting handles 29
Inventory 30
f201 lasers 30
Connecting 34
Cooling connections 34
PS-96 power supply connections 38
PS-96 AC input section 39
Control connections 42
Other connections 43
This is to certify that Firestar® f-Series lasers are guaranteed by SYNRAD to be free of all defects in materials and workmanship for a period of one year from the date of shipment. This warranty does not apply
to any defect caused by negligence, misuse (including environmental factors), accident, alteration,
or improper maintenance. We request that you examine each shipment within 10 days of receipt and
inform SYNRAD of any shortage or damage. If no discrepancies are reported, SYNRAD shall assume the
shipment was delivered complete and defect-free.
If, within one year from the date of shipment, any part of the f-Series laser should fail to operate, contact
the SYNRAD Customer Service department at 1.800.SYNRAD1 (outside the U.S. call 1.425.349.3500)
and report the problem. When calling for support, please be prepared to provide the date of purchase,
model number and serial number of the unit, and a brief description of the problem. When returning a
unit for service, a Return Authorization (RA) number is required; this number must be clearly marked on
the outside of the shipping container in order for the unit to be properly processed. If replacement parts
are sent to you, then you are required to send the failed parts back to SYNRAD for evaluation unless
otherwise instructed.
If your f-Series laser fails within the rst 45 days after shipment, SYNRAD will pay all shipping charges
to and from SYNRAD when shipped as specied by SYNRAD Customer Service. After the rst 45 days,
SYNRAD will continue to pay for the costs of shipping the repaired unit or replacement parts back to
the customer from SYNRAD. The customer, however, will be responsible for shipping charges incurred
when sending the failed unit or parts back to SYNRAD or a SYNRAD Authorized Distributor. In order to
maintain your product warranty and to ensure the safe and ecient operation of your f-Series laser, only
authorized SYNRAD replacement parts can be used. This warranty is void if any parts other than those
provided by SYNRAD are used.
SYNRAD and SYNRAD Authorized Distributors have the sole authority to make warranty statements
regarding SYNRAD products. SYNRAD and its Authorized Distributors neither assumes nor authorizes
any representative or other person to assume for us any other warranties in connection with the sale,
service, or shipment of our products. SYNRAD reserves the right to make changes and improvements
in the design of our products at any time without incurring any obligation to make equivalent changes
in products previously manufactured or shipped. Buyer agrees to hold SYNRAD harmless from any and
all damages, costs, and expenses relating to any claim arising from the design, manufacture, or use of
the product, or arising from a claim that such product furnished Buyer by SYNRAD, or the use thereof,
infringes upon any Patent, foreign or domestic.
SYNRAD f201 Operator’s Manual Version 3.2
vii
Page 10
Service & support information
SYNRAD worldwide headquarters are located north of Seattle in Mukilteo, Washington. U.S.A.
Our mailing address is:
Phone us at: 1.800.SYNRAD1 (1.800.796.7231)
Outside the U.S.: +1.425.349.3500
Fax: +1.425.349.3667
web: www.synrad.com
E-mail: synrad@synrad.com
Sales and Applications
SYNRAD Regional Sales Managers work with customers to identify and develop the best CO2
laser solution for a given application. Because they are familiar with you and your laser application, use them as a rst point of contact when questions arise. Regional Sales Managers also
serve as the liaison between you and our Applications Lab in processing material samples per
your specications. To speak to the Regional Sales Manager in your area, call SYNRAD at 1.800.
SYNRAD1.
Customer Service
For assistance with order or delivery status, service status, or to obtain a Return Authorization
(RA) number, contact SYNRAD at 1.800.SYNRAD1 and ask to speak to a Customer Service rep-
resentative, or you can email us by sending a message to customercare@synrad.com.
Technical Support
SYNRAD Regional Sales Managers are able to answer many technical questions regarding the
installation, use, troubleshooting, and maintenance of our products. In some cases, they may
transfer your call to a Laser, Marking Head, or Software Support Specialist. You may also e-mail
questions to the Technical Support Group by sending your message to customercare@synrad.
com.
Reference materials
Your Regional Sales Manager can provide reference materials including Outline & Mounting
drawings, Operator’s Manuals, Technical Bulletins, and Application Newsletters. Most of these
materials are also available directly from the SYNRAD web site at http://www.synrad.com.
SYNRAD f201 Operator’s Manual Version 3.2
viii
Page 11
EU headquarters
For assistance in Europe, contact SYNRAD® European subsidiary, SYNRAD Europe, at:
This chapter contains safety information that you will need to know prior to getting started.
yourself with all denitions and their signicance.
Hazard Information – includes equipment label terms and hazards, please familiarize
unsafe practices that could result in death, severe injury, or product damage.
pertain to disposal.
your f201 laser.
tion information.
General & Other Hazards – provides important information about the hazards and
Disposal – information on your f201 laser parts and/or components as they
Additional Safety Information – describes how to nd additional information about
Compliance – explains in the subsections therein applicable and appropriate regula-
Note: Read the entire safety section. This will ensure you are familiar with the hazards
and warnings prior to starting.
Warning
Serious
personal
injury
is Class 4 CO laser product emits invisible infrared laser
radiation in the 9.3–10.6 µm wavelength band.
Because direct or diuse laser radiation can inict severe
corneal injuries, always wear eye protection when in the same
area as an exposed laser beam.
Do not allow the laser beam to contact a person!
This product emits an invisible laser beam that is capable of
seriously burning human tissue.
Always be aware of the beam’s path and always use a beam
block while testing.
SYNRAD f201 Operator’s Manual Version 3.2
10
Page 13
Laser Safety
Person Lift
Hazard Information
Hazard information includes terms, symbols, and instructions used in this manual or on the
equipment to alert both operating and service personnel to the recommended precautions in
the care, use, and handling of Class 4 laser equipment.
Terms
Certain terms are used throughout this manual or on the equipment labels. Please familiarize
yourself with their denitions and signicance.
ParameterDescription
WARNING: Potential & Imminent hazards which, if not avoided,
DANGER: Hazards which, if not avoided, could result in minor or
moderate injury.
Two
Alerts operator of liing dangers.
CAUTION: Potential hazards or unsafe practices which, if not avoided,
may result in product damage.
Alerts operator of equipment dangers.
IMPORTANT NOTES & TIPS:
Important
Note:
Note:
Tip:
Content specic information and/or recommendations.
Figure 1-1 Labeling terms and denitions.
Warning
Serious
personal
injury
For laser systems being used or sold within the U.S.A., customers should refer to and follow the laser safety precautions
described American National Standards Institute (ANSI) document Z136.1-2014, Safe Use of Lasers.
For laser systems being used or sold outside the U.S.A., customers should refer to and follow the laser safety precautions
described in European Normative and International Electrotechnical Commission documents IEC/TR 60825-14:2014,
Safety of Laser Products – §14: A User’s Guide.
SYNRAD f201 Operator’s Manual Version 3.2
11
Page 14
Laser Safety
General hazards
Following are descriptions of general hazards and unsafe practices that could result in death,
severe injury, or product damage. Specic warnings and cautions not appearing in this section are found throughout the manual.
Danger
Serious
personal
injury
mal tissue, which may cause permanent damage.
This product is not intended for use in explosive, or potentially explosive, atmospheres!
Materials processing with a laser can generate air contaminants such as vapors, fumes,
and/or particles that may be noxious, toxic, or even fatal. Material Safety Data Sheets
(MSDS) for materials being processed should be thoroughly evaluated and the adequacy of provisions for fume extraction, ltering, and venting should be carefully considered. Review the following references for further information on exposure criteria:
Do not allow laser radiation to enter the eye by viewing direct
or reected laser energy.
CO laser radiation can be reected from metallic objects even
though the surface is darkened. Direct or diuse laser radiation
can inict severe corneal injuries leading to permanent eye
damage or blindness. All personnel must wear eye protection
suitable for CO radiation, e.g. 9.3–10.6 µm when in the same
area as an exposed laser beam.
Eye wear protects against scattered energy but is not intended to protect against direct viewing of the beam—never look
directly into the laser output aperture or view scattered laser
reections from metallic surfaces.
Enclose the beam path whenever possible. Exposure to direct
or diuse CO laser radiation can seriously burn human or ani-
ANSI Z136.1-2014, Safe Use of Lasers, section 7.3.
U.S. Government’s Code of Federal Regulations: 29 CFR 1910, Subpart Z.
Threshold Limit Values (TLV’s) published by the American Conference of Governmental
Industrial Hygienists (ACGIH).
It may be necessary to consult with local governmental agencies regarding restrictions
on the venting of processing vapors.
The use of aerosol dusters containing diuoroethane causes “blooming”, a condition
that signicantly expands and scatters the laser beam. This beam expansion can eect
mode quality and/or cause laser energy to extend beyond the connes of optical elements in the system, possibly damaging acrylic safety shielding. Do not use air dusters
containing diuoroethane in any area adjacent to CO laser systems because diuoroethane persists for long time periods over wide areas.
SYNRAD f201 Operator’s Manual Version 3.2
12
Page 15
Laser Safety
Light
Warning!
General hazards
F201 lasers should be installed and operated in manufacturing or laboratory facilities by
trained personnel only. Due to the considerable risks and hazards associated with the installation and operational use of any equipment incorporating a laser, the operator must follow
product warning labels and instructions to the user regarding laser safety. To prevent exposure to direct or scattered laser radiation, follow all safety precautions specied throughout
this manual and exercise safe operating practices per ANSI Z136.1-2014, Safe Use of Lasers at
all times when actively lasing.
Due to the specic properties of laser light, a unique set of safety hazards that dier from other light sources must be considered. Just like light, lasers can be reected, refracted, diracted
or scattered.
Never use organic
Serious
Personal
Injury
material or metals as
a beam blocker. There
are very few exceptions,
e.g. black anodized
metal such as aluminum
because this is non
reective surface.
Figure 1-2 Always wear safety glasses or protective goggles with side shields to reduce the
risk of damage to the eyes when operating the laser.
A CO2 laser is an intense energy source and will ignite most materials under the proper conditions. Never operate the laser in the presence of ammable or explosive materials, gases,
liquids, or vapors.
Warning
Serious
personal
injury
Safe operation of the laser requires the use of an external beam block to safely block the
beam from traveling beyond the desired work area. Do not place your body or any combustible object in the path of the laser beam. Use a water-cooled beam dump or power meter, or
similar non-scattering, noncombustible material as the beam block. Never use organic material or metals as the beam blocker; organic materials, in general, are apt to combust or melt and
metals act as specular reectors which may create a serious hazard outside the immediate
work area.
“Caution - The use of controls or adjustments or performance
of procedures other than those specied herein may result in
hazardous radiation exposure.”
SYNRAD f201 Operator’s Manual Version 3.2
13
Page 16
Laser Safety
Other hazards
The following hazards are typical for this product family when incorporated for intended use:
(A) risk of injury when lifting or moving the unit; (B) risk of exposure to hazardous laser energy
through unauthorized removal of access panels, doors, or protective barriers; (C) risk of exposure to hazardous laser energy and injury due to failure of personnel to use proper eye protection and/or failure to adhere to applicable laser safety procedures; (D) risk of exposure to
hazardous or lethal voltages through unauthorized removal of covers, doors, or access panels;
(E) generation of hazardous air contaminants that may be noxious, toxic, or even fatal.
Disposal
This product contains components that are considered hazardous industrial waste. If a situation occurs where the laser is rendered non-functional and cannot be repaired, it may be returned to SYNRAD who, for a fee, will ensure adequate disassembly, recycling and/or disposal
of the product.
Additional laser safety information
The SYNRAD web site http://www.synrad.com/LaserFacts/safetyinfo.html contains an online
laser safety handbook that provides information on (1) Laser Safety Standards for OEM’s/System Integrators, (2) Laser Safety Standards for End Users, (3) References and Sources, and (4)
Assistance with Requirements.
In addition, the Occupational Safety and Health Administration (OSHA) provides an online
Technical Manual located at http://www.osha.gov/dts/osta/otm/otm_iii/otm_iii_6.html. Section III, Chapter 6 and Appendix III are good resources for laser safety information.
Another excellent laser safety resource is the Laser Institute of America (LIA). Their comprehensive web site is located at http://www.lia.org.
SYNRAD f201 Operator’s Manual Version 3.2
14
Page 17
Laser Safety
F201 label locations
400 WATTS MAX
10200-10800 nm
MADE IN THE U.S.A.MADE IN THE U.S.A.
DANGER-INVISIBLE LASER
RADIATION. AVOID EYE OR SKIN
EXPOSURE TO DIRECT OR
SCATTERED RADIATION
CLASS 4 LASER PRODUCT
EN-60825-1, 2014
Keyswitch version
MODEL #: MODELNUMBER
SERIAL #: SERIALNUMBER
MFG Month DD, YYYY
Complies with IEC 60825-1 (1993) and 21 CFR 1040.10 except for deviations pursuant to
TESTED AT XX Volts
Specified by 21 CFR 1040.10 or IEC 60825-1
4600 Campus Place, Mukilteo WA 98275 | 425.349.3500
.
or
OEM version
MODEL #: MODELNUMBER
SERIAL #: SERIALNUMBER
MFG Month DD, YYYY
This laser component does not comply with standards for complete laser products as
TESTED AT XX Volts
Specified by 21 CFR 1040.10 or IEC 60825-1
4600 Campus Place, Mukilteo WA 98275 | 425.349.3500
AVOID EXPOSURE
Invisible laser radiation
is emitted from
this aperture.
.
Front
AVOID EXPOSURE
INVISIBLE LASER RADIATION
IS EMITTED FROM THIS
APERTURE
with Synrad's UC Series power controller. This signal
and facilitates breakdown and pulse to pulse fidelity.
keeps the plasma ionized during laser "low" periods
laser "low" periods. This is automatically provided
These RF Excited Lasers must be provided with a
equivalent drive signals are not used. See manual
for tickle pulse specifications or contact factory.
pre-ionizing "Tickle" signal during standby or
Damage or malfunction may occur if this or
This laser component does not comply with standards for complete
laser products as specified by 21 CFR 1040.10 or IEC 60825-1.
SYNRAD, Inc. 4600 Campus Place, Mukilteo WA 98275 425.349.3500
OCCUR IF COOLING WATER
SEE OPERATION MANUAL.
IS BELOW DEW POINT.
WATER DAMAGE CAN
CONDENSATION AN D
MODEL #: FSF100SD
SERIAL #: F100204082943
TESTED AT: 96V MFG: July 22, 2008
CAUTION
CAUTION
CAUTION
CONDENSATION AND
WATER DAMAGE CAN
OCCUR IF COOLING WATER
IS BELOW DEW POINT.
SEE OPERATION MANUAL.
WATER-COOLED VERSION ONLY
WARNING
Do not remove water fittings!
Do not lift or pull on water fittings!
This may cause misalignment or water leaks.
If a water leak is discovered, please contact
Synrad customer service immediately.
Top
10200-10800 NANOMETERS
400 WATTS MAX
EN-60825-1, 2007
DIRECT OR SCATTERED RADI ATION
AVOID EYE OR SKIN EXPOSURE TO
INVISIBLE LASER RADI ATION
CLASS 4 LASER PRODUCT
CAUTION
These RF Excited Lasers must be provided with a
pre-ionizing "Tickle" signal during standby or
laser "low" periods. This is automatically provided
with Synrad's UC Series power controller. This signal
keeps the plasma ionized during laser "low" periods
and facilitates breakdown and pulse to pulse fidelity.
Damage or malfunction may occur if this or
equivalent drive signals are not used. See manual
for tickle pulse specifications or contact factory.
Rear
WARNING
Do not remove water fittings!
Do not lift or pull on water fittings!
This may cause misalignment or water leaks.
If a water leak is discovered, please contact
Synrad customer service immediately.
This laser product is manufactured under
one or more of the following U.S. Patents:
Center for Devices and Radiological Health (CDRH) requirements.
Federal Communications Commission (FCC) requirements.
European Union (EU) requirements.
SYNRAD lasers are designed, tested, and certied to comply with certain United States (U.S.)
and European Union (EU) regulations. These regulations impose product performance requirements related to electromagnetic compatibility (EMC) and product safety characteristics for industrial, scientic, and medical (ISM) equipment. The specic provisions to which
systems containing f201 lasers must comply are identied and described in the following
paragraphs. Note that compliance to CDRH, FCC, and EU requirements depends in part on the
laser version selected—Keyswitch or OEM.
In the U.S., laser safety requirements are governed by the Center for Devices and Radiological
Health (CDRH) under the auspices of the U.S. Food and Drug Administration (FDA) while radiated emission standards fall under the jurisdiction of the U.S. Federal Communications Commission (FCC). Outside the U.S., laser safety and emissions are governed by European Union
(EU) Directives and Standards.
In the matter of CE-compliant laser products, SYNRAD assumes no responsibility for the
compliance of the system into which the product is integrated, other than to supply and/or
recommend laser components that are CE marked for compliance with applicable European
Union Directives.
Because OEM laser products are intended for incorporation as components in a laser processing system, they do not meet all of the Standards for complete laser processing systems as
specied by 21 CFR, §1040 or IEC 60825-1. SYNRAD assumes no responsibility for the compliance of the system into which OEM laser products are integrated.
Center for Devices and Radiological Health (CDRH)
requirements
Product features incorporated into the design of f201 lasers to comply with CDRH requirements are integrated as panel controls or indicators, internal circuit elements, or input/output
signal interfaces. Specically, these features include a lase and laser ready indicators, remote
interlock for power on/o, a laser aperture shutter switch, and a ve-second delay between
power on and lasing. Incorporation of certain features is dependent on the laser version (Keyswitch or OEM). Table 1, Class 4 safety features, indicates which features are available on f201
lasers, the type and description of the feature, and if the feature is required by CDRH regulations.
OEM models
F201 OEM lasers are OEM products intended for incorporation as components in laser pro-
cessing systems. As supplied by SYNRAD, these lasers do not meet the requirements of 21
CFR, Subchapter J without additional safeguards. In the U.S., the Buyer of these OEM laser
components is solely responsible for the assurance that the laser processing system sold to
an end user complies with all laser safety requirements before the actual sale of the system.
Under CDRH regulations, the Buyer must submit a report to the CDRH prior to shipping the
system. In jurisdictions outside the U.S., it is the sole responsibility of the Buyer of these OEM
SYNRAD f201 Operator’s Manual Version 3.2
18
Page 21
Laser Safety
components to ensure that they meet all applicable local laser safety requirements. In cases
where the Buyer is also the end-user of the OEM laser product, the Buyer/end-user must integrate the laser so that it complies with all applicable laser safety standards as set forth above.
Federal Communications Commission (FCC)
Requirements
The United States Communication Act of 1934 vested the Federal Communications Commission (FCC) with the authority to regulate equipment that emits electromagnetic radiation in
the radio frequency spectrum. The purpose of the Communication Act is to prevent harmful
electromagnetic interference (EMI) from aecting authorized radio communication services.
The FCC regulations that govern industrial, scientic, and medical (ISM) equipment are fully
described in 47 CFR, §18, §§C.
SYNRAD f201 lasers have been tested and found to comply by demonstrating performance
characteristics that have met or exceeded the requirements of 47 CFR, §18, §§C for Radiated
and Conducted Emissions.
FCC information to the user
Interference Potential
Note:
In our testing, SYNRAD has not discovered any signicant electrical interference traceable to f201 lasers.
System Maintenance
Ensure that all exterior covers are properly fastened in position.
Measures to Correct Interference
If you suspect that your laser interferes with other equipment, take the following steps to
minimize this interference:
1
2
The following FCC information to the user is provided to comply with the requirements
of 47 CFR, §18, §§ 213 Information to the user.
Use shielded cables to and from the equipment that is experiencing interfer-
ence problems.
Ensure that the laser is properly grounded to the same electrical potential as
the equipment or system it is connected to.
FCC caution to the user
The Federal Communications Commission warns the user that changes or modications of
the unit not expressly approved by the party responsible for compliance could void the user’s
authority to operate the equipment.
SYNRAD f201 Operator’s Manual Version 3.2
19
Page 22
Laser Safety
European Union (EU) requirements
RoHS compliance
SYNRAD f201 lasers meet the requirements of the European Parliament and Council Directive
2014/35/EU on the Restriction of the Use of Certain Hazardous Substances in Electrical and
Electronic Equipment that establishes maximum concentration values for certain hazardous
substances in electrical and electronic equipment.
Laser safety standards
Under the Low Voltage Directive, 2014/35/EU, the International Standard species the min-
imum documentation, marking and labeling for all laser products classied in accordance
with IEC 60825-1:2014 (hereafter referred to as EN 60825-1) including laser diodes and all laser
devices dened in ISO 11145 was developed to provide laser safety guidance and includes
clauses on Engineering Specications, Labeling, Other Informational Requirements, Additional Requirements for Specic Laser Products, Classication, and Determination of the Accessible Emission Level. To develop a risk assessment plan/laser safety program for users, see the
EN 60825-14:2004 Standard for the safety of laser products that includes clauses on Administrative Policies, Laser Radiation Hazards, Determining the MPE, Associated Hazards, Evaluating Risk, Control Measures, Maintenance of Safe Operation, Incident Reporting and Accident
Investigation, and Medical Surveillance.
OEM models
F201 OEM lasers are OEM products intended for incorporation as components in laser
processing systems. As supplied by SYNRAD, these lasers do not meet the requirements
of EN 60825-1 without additional safeguards. European Union Directives state that “OEM
laser products which are sold to other manufacturers for use as components of any
system for subsequent sale are not subject to this Standard, since the nal product will
itself be subject to the Standard.” This means that Buyers of OEM laser components are
solely responsible for the assurance that the laser processing system sold to an end-user complies with all laser safety requirements before the actual sale of the system. Note
that when an OEM laser component is incorporated into another device or system, the
entire machinery installation may be required to conform to EN 60825-1:2004, EN 602041:2016, Safety of Machinery; the Machinery Directive, 2006/42/EC; and/or any other applicable Standards and in cases where the system is being imported into the U.S., it must
also comply with CDRH regulations.
In cases where the Buyer is also the end-user of the OEM laser product, the Buyer/end-user must integrate the laser so that it complies with all applicable laser safety standards as
set forth above. Table 1, Class 4 safety features, summarizes f201 product features, indicating the type and description of features and whether those features are required by
European Union regulations.
Electromagnetic interference standards
SYNRAD f201 lasers have demonstrated performance characteristics that have met or exceeded the requirements of EMC Directive 2014/30/EU.
The European Union’s Electromagnetic Compatibility (EMC) Directive, 2014/30/EU, is the sole
Directive developed to address electromagnetic interference (EMI) issues in electronic equipment. In particular, the Directive calls out European Norm (EN) documents that dene the
emission and immunity standards for specic product categories. For f201 lasers, EN 610006-4:2018 denes radiated and conducted RF emission limits while EN 61000-6-2:2016 denes
immunity standards for industrial environments.
SYNRAD f201 Operator’s Manual Version 3.2
20
Page 23
Laser Safety
Table 1-1 Class 4 safety features.
Required by: Available on:
Feature Location / Description CDRH EN60825-1 OEM f201
Keyswitch1 Rear panel control Yes Yes No
On/O/Reset Keyswitch controls power to laser electronics.
Key cannot be removed from switch in the “On” position.
Shutter Laser control Yes Yes Yes
function Functions as a beam attenuator to disable RF driver/
laser output when closed.
Shutter Rear panel indicator (Blue) No No Yes
indicator Illuminates blue to indicate shutter is open.
Ready Rear panel indicator (Yellow) Yes Yes Yes
indicator Indicates that laser has power applied and is capable
of lasing.
Lase Rear panel indicator (Red) No No Yes
indicator Indicates that the laser is actively lasing. Lase LED
illuminates when the duty cycle of the Command signal
is long enough to produce laser output.
Five second Circuit element Yes No Yes
delay Disables RF driver/laser output for ve seconds after
Keyswitch is turned to “On” or remote reset/start pulse is
applied when Keyswitch is in “On” position.
Power fail Circuit element Yes Yes No
lockout1 Disables RF driver/laser output if input power is removed
then later reapplied (AC power failure or remote interlock
actuation) while Keyswitch is in “On” position.
Remote Rear panel connection Yes Yes Yes
Interlock Disables RF driver/laser output when a remote interlock
switch on an equipment door or panel is opened.
RemoteRear panel indicator (Green/Red) No No Yes
Interlock Illuminates green when Remote Interlock circuitry is closed.
Indicator Illuminates red when interlock circuitry is open.
Over Circuit element No No Yes
temperature Temperature shutdown occurs if temperature of the laser
protection tube rises above safe operating limits.
Temp Rear panel indicator (Green/Red) No No Yes
indicator Illuminates green when laser temperature is within
operating limits, changing to red when thermal limits
are exceeded.
Warning Laser exterior Yes Yes Yes
labels Labels attached to various external housing locations to
warn personnel of potential laser hazards.
1 Not available on f201 OEM lasers
SYNRAD f201 Operator’s Manual Version 3.2
21
Page 24
Laser Safety
When integrating SYNRAD f201 OEM lasers, the Buyer and/or integrator of the end system is
responsible for meeting all applicable Standards to obtain the CE mark. To aid this compliance
process, SYNRAD testing program has demonstrated that f201 lasers comply with the relevant
requirements of 2014/30/EU, the Electromagnetic Compatibility Directive, as summarized in
the table below.
EN 61010-1:2010 Safety Requirements for Electrical Equipment for Measure-
ment, Control, and Laboratory Use - Part 1: General
Requirements
EN 61000-6-4:2007 Radiated Emissions Group 1, Class A
EN 61000-6-4:2007 Conducted Emissions Group 1, Class A
EN 61000-6-2:2005 Electrostatic Discharge Immunity
EN 61000-6-2:2005 RF Electromagnetic Field Immunity
EN 61000-6-2:2005 Electrical Fast Transient/Burst Immunity
EN 61000-6-2:2005 Conducted RF Disturbances Immunity
After a laser or laser processing system has met the requirements of all applicable EU Directives, the product can bear the ocial compliance mark of the European Union as a Declaration of Conformity.
Use information in this chapter to prepare your Firestar f201 laser for operation. The
order of information presented in this chapter is the same as the order of tasks that
you will need to perform. The best way to get your laser ready for operation is to start
at Unpacking and work your way through Connecting.
This chapter contains the following information:
■Introduction – introduces the f201 laser, lists important features, and describes
Firestar nomenclature.
■Unpacking – provides important information about unpacking your f201 laser.
■Inventory – displays and describes all components shipped with your f201 laser.
■Mounting – describes how to attach your f201 laser to a mounting surface.
■Connecting – explains how to connect cooling tubing, power, and control cabling.
Caution
possible
equipment
damage
Caution
possible
equipment
damage
Operating the laser at coolant temperatures above 22 °C (72 °F)
may result in decreased performance and/or premature failure of
electronic components.
Operating the laser at coolant temperatures above 22 °C (72 °F)
may result in decreased performance and/or premature failure of
electronic components.
SYNRAD f201 Operator’s Manual Version 3.2
25
Page 28
getting started
Introduction
F201 lasers are part of a smaller-footprint lasers from SYNRAD featuring high power and excellent beam
quality. The circular beam provides more accurate cutting and faster processing speeds than competing
technologies. With an integrated RF power supply, meaning no external RF cables, the compact laser
housing mounts easily to atbed cutters, robotic arms, or gantry systems making integration into your
production line simple and fast.
f201 features include:
■ Compact resonator design ■ Color-coded LEDs mirror user outputs
■ Integrated RF power supply ■ “Industrial-strength” ±5 V to 24 VDC I/O
■ Built-in “tick le” generator ■ Keyswitch and OEM models available
Firestar nomenclature
Firestar f201 lasers are available only as OEM lasers; they do not include either a keyswitch or manual
shutter assembly. SYNRAD’s OEM lasers are primarily designed as components for integration into larger
processing systems by the Original Equipment Manufacturer (OEM) or System Integrator who bears the
responsibility for meeting the appropriate laser safety requirements for Class 4 laser systems.
Single tube
Firestar f201 lasers contain a single tube (optical resonator) within the laser chassis. These lasers incorporate a single set of controls and indicators as well as a single set of cooling ports.
Model numbers
The last three characters in the Firestar model number serve to designate the functional category,
cooling method, and model version. The functional category is indicated by an “S” (Switch-less) for OEM
models. The next letter indicates the cooling method: “W” for water-cooled units, “W” is omitted on f201
lasers since they are all water-cooled). The last letter in the model number indicates the current model
version beginning with “B”. For example, the model number. FSF201SC indicates a “C” version OEM
water-cooled f201 laser.
Important Note: All Firestar f201 lasers are water-cooled because of the integrated RF section’s heat
removal requirements. F201 lasers cannot be fan-cooled. Damage to your laser will
void your warranty.
SYNRAD f201 Operator’s Manual Version 3.2
26
Page 29
getting started
Unpacking
The Unpacking section includes subsections:
■ Incoming inspection
■ Packaging guidelines
■ Unpacking the laser
■ Removing the lifting handles
Incoming inspection
Upon arrival, inspect all shipping containers for signs of damage. If you discover shipping damage,
document the damage (photographically if possible), then immediately notify the shipping carrier and
SYNRAD.
The shipping carrier is responsible for any damage occurring during transportation from SYNRAD to
your receiving dock.
Packaging guidelines
Warning
possible
personal
Lifting or moving f201 lasers poses a potential for injury. Use appropriate lifting techniques and/or equipment to prevent a risk of
injury. In some cases, you may require assistance from additional
personnel to safely unpack and move this equipment.
injury
■ To prevent equipment damage or loss of small components, use care when removing packaging materi-
als.
■ After unpacking, review the Inventory section and verify that all components are on hand.
■ Do not lift or support the laser using the cooling ttings; lift the laser by the mounting feet or
baseplate only.
■ Save all shipping containers and packaging materials, including covers and plugs. Use these special-
ized packing materials when shipping the laser to another location.
■ When packing a laser for shipment, be sure to remove all accessory items not originally attached to
the laser including beam delivery components, cooling tubing, etc.
■ Refer to Packaging instructions drawings in the Technical Reference chapter for details on packag-
ing f201 lasers using SYNRAD-supplied shipping materials.
■ When shipping water-cooled lasers, remember to drain all cooling water from the laser and then cap
the open ttings to prevent debris from entering the coolant path.
■ F201 lasers and power supplies are heavy and awkward to move. Use appropriate lifting techniques,
additional personnel, and/or hoisting equipment to prevent a risk of injury when rigging this equipment.
SYNRAD f201 Operator’s Manual Version 3.2
27
Page 30
getting started
1
Unpacking
Unpacking the laser
To unbox the laser, refer to the gure below and perform the following steps:
5
2
6
FRAGILE
FRAGILE
FRAGILE
SAVE
THIS BO
X
6
3
4
Figure 2-1 Unpacking the laser.
1
Carefully cut the shipping strap wrapped around the container. This strap secures the upper lid to
the shipping crate.
2
Remove the upper lid and set it aside.
3
Remove the four polyurethane foam blocks that center the inner box within the outer crate.
4
Open the four aps of the inner box.
5
Remove the three foam blocks cradling the top of the laser. To remove the center block, rst slide
it forward several inches and then lift.
6
The laser is shipped with three lifting handles bolted to the mounting feet. Use these handle grips
to hoist or lift the laser from its shipping container. To prevent injury to personnel or damage to
the laser, always use a minimum of three people to handle the laser.
Note: Do not lift or support the laser using the cooling ttings. Lift the laser using the lifting handles,
mounting feet or baseplate only.
7
After removing the laser, pull out all the accessory items packaged with the laser.
SYNRAD f201 Operator’s Manual Version 3.2
28
Page 31
getting started
Remove these
bolts
Unpacking
Removing the lifting handles
Once you have placed the laser in its nal mounting location, refer to the gure below and perform the
following steps to remove the lifting attachments.
10200-10800 NANOMETERS
1400 WATTS MAX
EN-60825-1, 1993
AVOID EYE OR SKIN EXPOSURE TO
INVISIBLE LASER RADIATION
DIRECT OR SCATTERED RADIATION
CLASS 4 LASER PRODUCT
CONDENSATION AND
CAUTION
OCCUR IF COOLING WATER
WATER DAMAGE CAN
IS BELOW DEW POINT.
SEE OPERATION MANUAL.
This laser component does not comply with standards for complete
laser products as specified by 21 CFR 1040.10 or IEC 60825-1.
MODEL #: FSF400SC
SERIAL #: F400204082943
TESTED AT: 96V MFG: July 22, 2008
SYNRAD, Inc. 4600 Campus Place, Mukilteo WA 98275 425.349.3500
Figure 2-2 Removing lifting handles.
1
Locate the 8 mm hex wrench included with the accessory items.
2
Remove the M10 × 1.5 metric capscrews from the three locations shown in the above gure.
3
Retain the three M10 × 1.5 capscrews. Use these bolts for mounting if metric fasteners are required.
4
Store the lifting attachments so that they can be reinstalled if the laser is moved to another loca-
Mounting Bolts ..............................................3/6 Final Test Report (not shown) .................. 1
SYNRAD CO2 Welcome Flier ...................... 1
f201 contents description
Each item listed in the table above is described below:
DC Power Cable – carries DC power from the DC power supply to your Firestar laser.
Quick Start Plug – connects to Firestar’s User I/O connector. Jumpers are built into the plug to enable
Firestar’s shutter and remote interlock circuits for initial start-up and testing.
BNC Control Cable (f201 only) – Coaxial cable carries the PWM Command signal from the UC-2000
Controller to the laser’s Quick Start Plug.
Cooling Tubing – Carries cooling water from the chiller to the laser and back. This black polyethylene
tubing is 1/2-inch O.D. by 30 feet and must be cut to length.
Mounting Bolts –Fasten our laser to your mounting surface. Firestar f201 lasers include three metric (M8
× 1.25 × 30 mm) and three standard (5/16–18 × 1.25" UNC) capscrews.
SYNRAD CO2 Welcome Flier – Contains an f201 customer welcome that provides where to nd, operation, and maintenance information for your f201 single-tube laser on our website.
Chiller Fittings – Adapt 1/2-inch O.D. tubing to 3/8" NPT threads for connection to a chiller. The chiller’s
inlet and outlet ports must be equipped with, or be adapted to, 3/8" NPT female threads.
Spherical Washer Sets (not shown) – used in conjunction with Firestar f201 mounting bolts to align the
laser to your mounting surface. Shipped with f201 lasers only.
Cooling Kit – adapts the laser’s straight 1/2-inch coolant ttings to 90° adaptor ttings for either 1/2-inch
standard or 12-mm metric cooling tubing.
Final Test Report (not shown) – contains data collected during the laser’s nal pre-shipment test.
SYNRAD f201 Operator’s Manual Version 3.2
31
Page 34
getting started
A
A
-
Mounting
f201 mounting
F201 mounting brackets are threaded to accept either metric or standard fasteners.
Important Note: As shown in the gure below, DO NOT use more than one fastener per mounting
location. Using two or more fasteners per mounting tab will cause damage to the
laser; use only one fastener per location to mount your f201 laser.
Do not use more than
one fastener per mount
ing foot. Doing so will
damage the laser
Fastening f201 lasers from above
To fasten your f201 laser to a mounting surface from above, perform the following steps:
1
Refer to the appropriate outline and mounting drawing for dimensions and then drill and tap
three mounting holes into your mounting surface. For SAE installations, drill and tap three 5/16–18
(f201).
These hole locations (referenced by Note 2 on the O & M drawings) should correspond to the thru
holes labeled “A” as shown in the gure below.
2
Assemble three spherical washer sets as detailed in the Outline and Mounting drawing.
3
Place the spherical washer sets over the threaded holes in your mounting surface.
4
Carefully position the laser so that each thru hole in the baseplate aligns with a spherical washer
and threaded mounting hole.
5
Insert the appropriate metric or standard capscrews, included in the laser’s ship kit, through the
baseplate, through the spherical washers, and into the threaded holes of the mounting surface
(see gure below). Turn the screws by hand until the threads engage.
6
Evenly tighten all three capscrews. For M8 capscrews, torque to 12 N m; for M10 screws, torque to
24 N m. When using 5/16" screws, torque to 13 ft lb; for 3/8" capscrews, tighten to 23 ft lb.
damage the laser
Do not use more than
ing foot. Doing so will
CAUTION
CONDENS ATION AN D
WATER DAMAGE CA N
IS BELOW DEW POIN T.
OCCUR IF COOLING WATER
sense fault conditions that could potentially
damage the laser’s electronic circuit boards.
The RF Drive circuit in this laser is designed to
On rare occasions, the laser may shutdown during
one fastener per mount-
Insert metric or UNC
fastener thru mounting
set into threaded metric
or UNC hole in laser mount
plate and spherical washer
OR
mounting plate
fastener thru laser
mount and spherical
Insert metric or UNC
washer set into threaded
MODEL #: FSF400SD
SERIAL #: F400204082943
TESTED AT: 96V MFG: July 22, 2008
SYNRAD, Inc. 4600 Campus Place, Mukilteo WA 98275 425.349.3500
laser products as specified by 21 CFR 1040.10 or IEC 60825-1.
This laser component does not comply with standards for complete
SEE OPER ATION MA NUAL.
re-apply DC power.
problem in the laser’s control circuit.
SYNRAD, Inc. as this may indicate a serious
the start-up sequence. When this happens, the
power from the laser, wait 30 seconds, and then
If the Ready indicator continues to flash, contact
Ready indicator will flash a series of three blinks,
pause, and then repeat. If this occurs, remove DC
CLASS 4 LASER PRODUCT
INVISIBLE LASER RADI AT IO N
AVOID EYE OR SKIN EXPOSURE TO
DIRECT OR SC ATTERED RADI ATION
EN-60825-1, 2007
1400 WATTS MA X
10200-10800 NANOMETERS
Top View
A
Figure 2-4 Fastening the f201 from below.
SYNRAD f201 Operator’s Manual Version 3.2
32
Cooling ttings
removed for clarity
Page 35
getting started
washer set into threaded
M
M
M
S
S
Mounting
f201 mounting
Insert metric or SAE
fastener thru laser
mount and spherical
mounting plate
Figure 2-5 Detail of fastening into mounting surface.
Fastening f201 lasers from below
To fasten your f201 laser to a mounting surface from below, perform the following steps:
1
Refer to the appropriate outline and mounting drawing for dimensions and then drill three 8.5
mm (f201) holes into your mounting surface. For SAE installations, drill three 0.332" (f201) holes
into your mounting surface.
These hole locations (referenced by Note 1 or Note 3 on the O & M drawings) should correspond
to the threaded holes labeled “M”, for metric, or “S”, for standard, as shown in the following gure.
damage the laser
Do not use more than
ing foot. Doing so will
one fastener per mount-
C A UTION
CONDENS A TION AN D
W A TER D AM A GE CA N
The RF Drive circuit in this laser is designed to
OCCUR IF COOLING W A TE R
sense fault conditions that could potentially
damage the laser’s electronic circuit boards.
S
Insert metric or UNC
fastener thru mounting
plate and spherical washer
OR
fastener thru laser
mount and spherical
Insert metric or UNC
MODEL #: FSF400SD
SERIAL #: F400204082943
TESTED AT: 96V MFG: July 22, 2008
SYNRAD, Inc. 4600 Campus Place, Mukilteo WA 98275 425.349.3500
laser products as specified by 21 CFR 1040.10 or IEC 60825-1.
This laser component does not comply with standards for complete
IS BEL O W DEW POIN T .
SEE OPER A TION MAN U AL .
re-apply DC power.
problem in the laser’s control circuit.
SYNRAD, Inc. as this may indicate a serious
the start-up sequence. When this happens, the
power from the laser, wait 30 seconds, and then
If the Ready indicator continues to flash, contact
Ready indicator will flash a series of three blinks,
pause, and then repeat. If this occurs, remove DC
On rare occasions, the laser may shutdown during
set into threaded metric
washer set into threaded
or UNC hole in laser mount
mounting plate
Top View
CLASS 4 LASER P R ODUC T
INVISIBLE LASER RADI A TIO N
A V OID EYE OR SKIN EXPOSURE T O
DIRECT OR SC A TTERED RADI A TIO N
EN-60825-1, 2007
1400 W A TTS MA X
10200-10800 NANOMETERS
(f400 model shown)
Cooling fittings
removed for clarity
Figure 2-6 Fastening f201 from below.
2
Assemble three spherical washer sets as detailed in the outline and mounting drawing.
3
Place the spherical washer sets over the thru holes in your mounting surface.
4
Carefully position the laser so that each threaded hole in the baseplate (labeled “M” or “S” in the
prior gure) aligns with a spherical washer and thru hole in the mounting surface.
5
Insert the appropriate metric or standard capscrews, included in the laser’s ship kit, through the
mounting surface and spherical washers into the threaded holes of Firestar’s baseplate (see Figure
1-11). Turn the screws by hand until the threads engage.
6
Evenly tighten all three capscrews. For M8 capscrews, torque to 12 N m; for M10 capscrews,
tighten to 24 N m. When using 5/16" capscrews, torque to a value of 13 ft lb; for 3/8" capscrews,
tighten to 23 ft lb.
SYNRAD f201 Operator’s Manual Version 3.2
33
Page 36
getting started
Connecting
The Connecting section includes subsections:
■ Cooling connections
■ PS-96 power supply connections
■ Control connections
■ Other connections
Cooling connections
Read Guidelines for cutting and installing tubing before installing any cooling tubing and then make sure to
connect the cooling system exactly as described for your particular laser.
Guidelines for cutting and installing tubing
■ Cut tubing lengths generously to allow for trimming.
■ Cut tubing squarely; diagonal cuts may not seal properly. Trim away burrs if the cut is “ragged”.
■ Avoid excessive stress on ttings; create gentle bends when routing tubing close to connectors. Exces-
sive stress from sharp bends will compromise the sealing properties of the tting.
■ Never allow the tubing to kink, since kinking severely restricts coolant ow.
■ Push tubing completely into the tting, then pull the tubing to verify that it is locked into place. Tub-
ing extends into the tting approximately 7/8 of an inch.
■ If tubing must be disconnected from a tting, rst push and hold the tubing slightly into the tting.
Next push the white tting ring evenly towards the tting, and then pull the tubing free.
■ After disconnecting tubing from a tting, trim 12.7 mm (1/2") from its end before reconnecting. Trim-
ming the end of the tubing before reconnecting the tting provides an undisturbed sealing surface.
Laser cooling ttings
If your integrated laser application uses metric cooling tubing, we recommend the installation of tubing
adaptors to convert the laser’s existing WATER IN/WATER OUT ttings from 1/2-inch standard to 12-mm
metric tubing. The ship kit sent with your f201 laser includes a cooling kit containing two each 1/2-inch
male to 1/2-inch female 90° adaptors and two each 1/2-inch male to 12-mm female 90° adaptors. If
required, many tubing and tting manufacturers can supply 1/2-inch to 12-mm straight ttings.
Connecting
Caution
Firestar f201 lasers incorporate special ttings that accept only standard 1/2" O.D. cooling tubing.
possible
equipment
DO NOT REMOVE OR ADJUST THESE FITTINGS!
damage
Each laser is shipped with a cooling kit that includes 90° push-on
tubing connectors to adapt the brass ttings to your choice of standard (1/2-inch) or metric (12-mm) cooling tubing.
SYNRAD f201 Operator’s Manual Version 3.2
34
Page 37
getting started
Mounting
Chiller preparation guidelines
■The ship kit includes chiller ttings with 3/8" NPT male threads. If these ttings will not connect to
your chiller then you must provide ttings that will adapt the laser’s 1/2" O.D. polyethylene cooling
tubing to your chiller’s Inlet and Outlet ports.
■Because Firestar’s cooling tubing is supplied in inch sizes, do not use metric tubing ttings unless you
have installed the appropriate inch-to-metric tubing adaptors. The use of metric ttings on inch size
tubing will lead to coolant leaks or may allow the pressurized tubing to blow o the tting.
Coolants
SYNRAD recommends the laser’s cooling uid contain at least 90% distilled water by volume. In closedloop systems, use a corrosion inhibitor/algaecide such as Optishield® Plus or equivalent as required.
Avoid glycol-based additives because they reduce the coolant’s heat capacity and high concentrations
may aect power stability. For SYNRAD lasers, the minimum coolant setpoint is 18 °C (64 °F) so glycol is
not necessary unless the chiller is subjected to freezing temperatures. If tap water is used, chloride levels
should not exceed a concentration of 25 parts per million (PPM) and total hardness should be below 100
PPM. Install a lter on the chiller’s return line and inspect frequently. f201 lasers use the following wetted materials in the coolant path—aluminum, brass, copper, Delrin®, PBT, polyethylene, stainless steel,
and Viton®.
Note: DO NOT use de-ionized (DI) water as a coolant. DI water is unusually corrosive and is not recom-
mended for mixed material cooling systems.
Setting coolant temperature
Choosing the correct coolant temperature is important to the proper operation and longevity of your laser.
When coolant temperature is lower than the dew point (the temperature at which moisture condenses out
of the surrounding air), condensation forms inside the laser housing leading to failure of laser electronics as
well as damage to optical surfaces.
The greatest risk of condensation damage occurs when the laser is in a high heat/high humidity environment and the chiller’s coolant temperature is colder than the dew point of the surrounding air or when the
system is shut down, but coolant continues to ow through the laser for extended periods of time.
The chiller’s temperature setpoint must always be set above the dew point temperature. In cases where this
is not possible within the specied coolant temperature range of 18 °C to 22 °C (64 °F to 72 °F), then the
following steps MUST be taken to reduce the risk of condensation damage.
■ Use the Gas Purge port (f201) to introduce nitrogen or dry, ltered air into the laser housing.
■ Air-condition or dehumidify the room or the enclosure containing the laser.
■ Increase coolant ow by an additional 3.8 lpm (1.0 GPM). Do not exceed a coolant pressure of 414
kPa (60 PSI).
Note: To reduce problems associated with condensation, Firestar f201asers can operate up to 28 °C.
However, these elevated cooling temperatures may result in decreased laser performance and/
or reduced laser lifetime.
SYNRAD f201 Operator’s Manual Version 3.2
35
Page 38
getting started
The following table provides dew point temperatures for a range of air temperature and relative humidity values. Remember that the laser’s coolant temperature must be set above the dew point temperatures given in the chart; however, for best results and performance, use a coolant temperature in the
range of 18–22 °C (64–72 °F) when possible.
To use the table above, look down the Air Temp column and locate an air temperature in Fahrenheit or
Celsius (°C values are shown in parentheses) that corresponds to the air temperature in the area where
your laser is operating. Follow this row across until you reach a column matching the relative humidity in your location. The value at the intersection of the Air Temp and Relative Humidity columns is the
Dew Point Temperature in °F (or °C). The chiller’s temperature setpoint must be set above the dew point
temperature. For example, if the air temperature is 85 °F (29 °C) and the relative humidity is 60%, then
the dew point temperature is 70 °F (21 °C). Adjust the chiller’s temperature setpoint to 72 °F (22 °C) to
prevent condensation from forming inside the laser.
SYNRAD f201 Operator’s Manual Version 3.2
36
Page 39
getting started
Connecting
f201 cooling tubing connections
To connect cooling tubing to your f201 laser, refer to the gure below and perform the following steps.
The numbered items that follow correspond to the step numbers in the following procedure.
Chiller
OUTLET
INLET
20A FAST
FUSE
DC POWER
96 VDC 18A MAX
WATER OUT
Remote
Interlock
WATER IN
OFF/RESET
Lase
Shutter
Ready
Temp
ON
USERI/O
Figure 2-7 F201 ooling connections.
1
Cut and connect a length of cooling tubing to t between the chiller’s Outlet port and the
WATER IN port on the rear of your f201 laser.
2
Cut and connect a length of cooling tubing to t between the WATER OUT port on the rear of the
laser and the chiller’s Inlet port.
3
Turn on the chiller and adjust the temperature setpoint to 18 °C to 22 °C. Regulate coolant ow to
9.5 lpm (2.5 GPM) for f201 lasers at less than 414 kPa (60 PSI) of pressure.
4
Closely examine all cooling connections and verify that there are no leaks.
1
Locate the 1/2-inch O.D. cooling tubing in the Ship Kit and cut four lengths (12–18 inches long) to
t partway between your chiller and the rear of the laser.
2
Select two lengths of tubing and connect one end of each piece to the laser’s WATER IN ports.
3
Connect the other end of each length of tubing to the “Y” tting (included in the Ship Kit).
Caution
possible
equipment
damage
Inlet cooling water temperature must always be maintained above
the dew point to prevent condensation and water damage to your
Firestar laser.
SYNRAD f201 Operator’s Manual Version 3.2
37
Page 40
getting started
4
Cut and connect a length of tubing to t between the chiller’s Outlet port and the center of the “Y”
tting.
5
Select the other two lengths of previously cut 1/2-inch cooling tubing and connect one end of
each piece to the laser’s WATER OUT ports.
6
Connect the other end of each length of tubing to the other “Y” tting.
7
Cut and connect a length of tubing to t between the chiller Inlet and the center of the “Y” tting.
8
Turn on the chiller and adjust the temperature setpoint to 18 °C to 22 °C. Regulate coolant ow to
15.1 lpm (4.0 GPM) of pressure.
9
Closely examine all cooling connections, assure there are no leaks.
Connecting
PS-96 power supply connections
The following procedures describe how to connect the PS-96 DC power supply to your f201 laser.
When connecting other types of DC power sources, remember to follow proper Lockout/Tagout procedures to ensure that AC power is de-energized, and then connect input AC power per the manufacturer’s instructions. Firestar f201 lasers require a DC power source capable of supplying 96 VDC.
Note: The negative (–) side of the DC input to the laser is internally connected so that the laser chassis
serves as DC power ground. You should isolate the laser’s DC power supply so that the only grounded
connection is at the laser. Alternatively, you can mount the laser chassis on an insulating pad or lm in
order to electrically isolate the laser when other equipment is grounded to the laser’s DC power supply.
Warning
serious
personal
injury
Caution
possible
equipment
Hazardous DC voltages exist on DC power supply output terminals when the power supply is energized. Contacting energized
terminals may result in serious personal injury or death. Protect all
96 VDC connections from incidental contact in accordance with
National, State, and Local requirements for electrical insulation and
labeling.
Do not reverse polarity when connecting the DC Power cable to
your DC power source. Reversed DC polarity may damage the laser’s internal RF power supply. Carefully follow the directions below
to ensure that DC Power cable leads are properly connected to the
correct DC output terminals.
damage
Caution
possible
equipment
Do not plug and unplug the DC Power cable on the rear of the laser
to switch DC power to the laser. Switching power under load causes
arcing that will damage the connector and control board. To properly cycle power to the laser, switch the AC power source controlling the PS-96 DC power supply.
damage
SYNRAD f201 Operator’s Manual Version 3.2
38
Page 41
getting started
DC Power cable
To connect the DC Power cable to your f201 laser, refer to the following gure and perform the steps
described below.
DC Power cable—f201 lasers
1 Verify that input AC voltage to the DC power supply is physically locked out or disconnected.
2 Locate the f201 DC Power cable. The DC Power cable is 5 meters (16 feet) long with a plug on one
end and ve (5) connection lugs on the other. This cable has a minimum bend radius of 87.4 mm
(3.44").
3 Carefully remove the plastic terminal guard from the output terminal block.
Important Note: Always re-install the terminal guard to protect users against inadvertent contact with
hazardous DC voltages on the connection lugs and DC output terminals.
4 On the lug end of the cable, attach the two black wires with the “0 VDC” marker to the negative (–)
96 VDC output terminal on the DC power supply. Tighten the terminal nut to a maximum torque
of 4.0 N m (35 in lb).
PS-96 AC input section
Three-phase operation
SYNRAD’s PS-96 DC power supply is capable of supplying 96 VDC at 47 amperes maximum.
Three-phase AC input power requirements for the PS-96 supply are 105–264 VAC, 3Ø (three
phase with ground phase to phase), 23 A maximum per phase, 47–63 Hz. The following table
provides recommendations for three-phase input wiring and fusing.
Caution
possible
personal
injury
All AC input wiring and fusing to the DC power supply must be
sized and connected in accordance with applicable local, state, and
national requirements.
Local, state, and national code requirements (like NEC, CSA, and IEC
60364) supersede any recommendations provided in this manual.
(recommended)
Table 2-4 Three-phase AC electrical recommendations.
Parameter Recommendation
Input voltage 105–264 VAC, 3Ø
Input current, max. 23 Amperes (per phase)
Wire gauge 10 AWG
Fuse/circuit breaker 30 Amps
To connect three-phase AC input power to the PS-96 power supply, refer back to the prior gure and
perform the following steps:
SYNRAD f201 Operator’s Manual Version 3.2
39
Page 42
getting started
+96 VDC Output
Connecting
Caution
possible
equipment
damage
Caution
possible
personal
injury
VAC
Do not plug and unplug the DC Power cable on the rear of the laser
to switch DC power to the laser. Switching power under load causes
arcing that will damage the connector and control board. To properly cycle power to the laser, switch the AC power source controlling the PS-96 DC power supply.
All AC input wiring and fusing to the DC power supply must be
sized and connected in accordance with applicable local, state, and
national requirements.
Local, state, and national code requirements (like NEC, CSA, and IEC
60364) supersede any recommendations provided in this manual.
Chassis Ground
– DC Output
Figure 2-8 PS-96 interface.
Note: Because AC input connections and requirements vary from facility to facility, cus-
tomers must provide the AC power cable or wiring.
1 Follow your facility’s Lockout/Tagout procedures and verify that input AC power to the DC power
supply is physically locked out or disconnected.
2 Connect the safety ground (earth) wire, typically green, to the input terminal labeled “GND”.
3 Connect the rst AC three-phase hot wire, typically black, to the input terminal labeled “L1/L”.
4 Connect the second three-phase hot wire, typically red, to the input terminal labeled “L2/L”.
5 Connect a jumper wire (the same wire gauge as the other AC wires) between the terminals labeled
“L2/L” and “L2/N”.
6 Connect the third AC hot wire, typically blue, to the input terminal labeled “L3/N”.
Single-phase input power requirements for SYNRAD PS-96 DC power supply are 208–264 VAC, 1Ø
(single phase), 20 A maximum, 47–63 Hz. Table below provides recommendations for single-phase input
wiring and fusing.
SYNRAD f201 Operator’s Manual Version 3.2
40
Page 43
getting started
Connecting
Single-phase operation
Table 2-5 Single-phase AC electrical recommendations.
Parameter Recommendation
Input voltage 208–264 VAC, 1Ø
Input current, max. 20 Amperes
Wire gauge 10 AWG
Fuse/circuit breaker 30 Amps
There are two possible circuit congurations for single-phase connections: (1) one circuit where the
voltage between hot conductors measures 208–264 VAC (recommended), or (2) an alternate case using
two individual circuits where the voltage between the hot and neutral conductors from each individual
circuit measures 120–264 VAC. To connect single-phase AC input power to the PS-96 power supply, refer
back to prior gures and perform the following steps:
Note: Because AC input connections and requirements vary from facility to facility, cus-
tomers must provide the AC power cable or wiring.
One single-phase circuit (208–264 VAC)
1 Follow your facility’s Lockout/Tagout procedures and verify that input AC power to the DC power
supply is physically locked out or disconnected.
2 Connect the safety ground (earth) wire, typically green, to the input terminal labeled “GND”.
3 Connect one AC hot wire, typically black, to the input terminal labeled “L1/L”.
4 Connect a jumper wire (the same wire gauge as the other AC wires) between the terminals labeled
“L1/L” and “L2/L”.
5
Connect the other hot wire, typically black, to the input terminal labeled “L2/N.
6
Connect a jumper wire (the same wire gauge as the other AC wires) between the terminals labeled
“L2/N” and “L3/N”.
Note: The neutral white, typically white, is not used in this wiring scheme.
Two individual single-phase circuits (120–264 VAC)
1 Follow your facility’s Lockout/Tagout procedures and verify that input AC power to the DC power
supply is physically locked out or disconnected.
2 Connect the safety ground (earth) wires from both circuit 1 and circuit 2, typically green, to the
input terminal labeled “GND”.
3 Connect the AC hot wire from circuit 1, typically black, to the input terminal labeled “L1/L”.
4 Connect the return (neutral) wire from circuit 1, typically white, to the input terminal labeled
“L2/N”.
5 Connect the hot wire from circuit 2, typically black, to the input terminal labeled “L2/L”.
6 Connect the return (neutral) wire from circuit 2, typically white, to the input terminal labeled
“L3/N”.
SYNRAD f201 Operator’s Manual Version 3.2
41
Page 44
getting started
Control connections
All control connections to Firestar f201 lasers are made through the 15-pin User I/O connector on
the rear panel. The User I/O port receives power commands from SYNRAD’s UC-2000 Universal Laser
Controller or FH Series Marking Head and also serves as the connection point for auxiliary signals between the laser and any parts handling, automation, or monitoring equipment.
UC-2000 Universal Laser Controller
SYNRAD recommends the use of a UC-2000 Universal Laser Controller to generate pulse width modulated (PWM) Command signals that control the laser’s output power. To connect a UC-2000 Controller
(available separately), perform the steps described in the appropriate f201 laser section below:
Note: F -Series lasers can also be controlled from an alternate user-supplied Command signal source.
Refer to Controlling laser power in the Technical Reference chapter for control signal descriptions and refer to User I/O connections, also in the Technical Reference chapter, for signal specications and connection details.
f201 lasers
1 Disconnect DC power from the laser.
2 Locate the Quick Start Plug in the ship kit.
3 Connect the Quick Start Plug to the User I/O connector on the rear of the f201 laser.
4 Attach the BNC connector on the end of the UC-2000’s Power/Control cable to the BNC connector
on the rear of the Quick Start Plug.
5 Connect the miniature DC power plug on the UC-2000’s Power/Control cable to the miniature con-
nector on the cable from the UC-2000’s wall plug transformer.
6 Connect the mini-DIN connector on the other end of the UC-2000’s Power/Control cable to the
Laser connector on the UC-2000’s rear panel.
7 Plug the compact transformer into any 100–240 VAC, 50–60 Hz outlet.
Connecting
Quick Start Plug
In order for your f201 laser to properly operate, several input signals must be applied to the DB-15 User
I/O connector before lasing is enabled. Voltage must be applied to Remote Interlock (Pin 3) and Shutter
Open Request (Pin 10) inputs before the laser will re. In applications where Firestar is integrated into
an automated system and safety interlocks are required, these input signals must be provided by the
customer’s control system. The Quick Start Plug included in the ship kit has factory-installed shorting
jumpers wired into it to enable these inputs. Connect the Quick Start Plug to the User I/O connector
when performing initial start-up and testing of your Firestar laser.
For further information about the User I/O connector, see User I/O connections in the Technical Reference chapter for User I/O pinouts and signal descriptions. See Integrating Firestar safety features, also in
the Technical Reference chapter, for detailed instructions on integrating your lasers keyswitch, shutter,
and remote interlock functions with automated control systems.
SYNRAD f201 Operator’s Manual Version 3.2
42
Page 45
getting started
Other connections
Connecting
Warning
serious
personal
injury
Caution
possible
equipment
damage
The use of the Quick Start Plug bypasses the laser’s safety interlock
function, potentially exposing personnel in the area to invisible
infrared laser radiation.
Because this plug jumpers Remote Interlock and Shutter Open
Request signals, the laser will re immediately on application of
a PWM Command signal. Your integrated control system should
provide interlock and shutter signals directly to the DB-15 User I/O
connector only after safe operating conditions are established.
The Quick Start Plug is intended only for initial testing and troubleshooting by qualied personnel. In normal operation, the laser’s
Remote Interlock input should be connected to the machine’s
safety interlock circuitry.
Turn o DC power before installing or removing any plug or cable
from the User I/O connector. Ensure that user connections are made
to the appropriate pins and that the appropriate signal levels are
applied. Failure to do so may damage the laser.
Gas Purge connector
A gas purge is highly recommended when operating the laser in dirty or dusty environments. Purging the
laser creates a positive pressure inside the laser housing that prevents dirt and debris from accumulating on
optical surfaces including beam combiner, telescope, and turning optics. In condensing atmospheres, a gas
purge helps to reduce the potential for condensation damage.
To connect the f201 gas purge port, perform the steps described below:
1
Connect nitrogen or breathing-grade air to the Gas Purge connector using 1/4-inch plastic tubing.
2
Push the tubing completely into the tting and then pull the tubing lightly to ensure that it is locked
into place.
Note: To disconnect gas purge tubing, rst push and hold the tubing slightly into the tting. Next push
the white tting ring evenly towards the tting, and then pull the tubing free.
3
Set a purge pressure between 13.8–34.5 kPa (2–5 PSI). This provides just enough positive airow
to prevent dust from entering the laser. If a owmeter is available, set a ow rate of 0.85–1.7 m3/hr
(30–60 SCFH, Standard Cubic Feet per Hour) at a pressure not to exceed 34.5 kPa (5 PSI).
The Gas Purge connector on your f201 laser must be connected to a source of nitrogen or clean, dry air
only; do not use any other gases for purging. Purge gas specications are listed in the table below.
SYNRAD f201 Operator’s Manual Version 3.2
43
Page 46
getting started
Gas Purge connector
Table 2-6 Purge gas specications.
Purge Gas Specication
Nitrogen High Purity Grade > 99.9500% purity; ltered to ISO Class 1 particulate level
Air Breathing Grade > 99.9996% purity; ltered to ISO Class 1 particulate level
Air Compressed Instrument-grade air ltered and dried to ISO 8573-1:2010
Class 1, 2, 1 (< 10 1.0–5.0 µm particles/m3; < –40 °F dew
point; < 0.01 mg/m3 oil vapor)
Caution
possible
equipment
damage
Do not exceed a gas purge pressure of 34.5 kPa (5 PSI). Excessive
pressure may damage the purge assembly or other internal laser
components.
Do not use argon as a purge gas. Use only nitrogen or clean, dry air
as described in Table 1-6, Purge gas specications.
SYNRAD f201 Operator’s Manual Version 3.2
44
Page 47
operation
operation
3
Use information in this chapter to familiarize yourself with your lasers controls and
indicators and to begin operating the laser.
This chapter contains the following information:
■ Controls and indicators – displays and describes exterior controls and indicators on
f201 lasers.
■ Initial start-up – explains how to start your f201 laser while verifying proper opera-
tion.
Danger
serious
personal
injury
Warning
serious
personal
injury
Warning
serious
personal
injury
This Class 4 laser product emits invisible infrared laser radiation
in the 10.6 µm CO2 wavelength band. Since direct or diuse laser
radiation can inict severe corneal injuries, always wear eye protection when in the same area as an exposed laser beam. Do not allow
the laser beam to contact a person. This product emits an invisible
laser beam that is capable of seriously burning human tissue.
Always be aware of the beam’s path and always use a beam block
while testing.
On Firestar f201 OEM lasers, remote interlock faults are not latched.
Clearing the fault condition re-enables the Ready indicator and the
laser will re after the ve-second delay provided that the Shutter indicator is lit and a PWM Command signal is applied. Because
exposure to 10.6 µm CO2 laser radiation can inict severe corneal injuries and seriously burn human tissue, the OEM or System Integrator must ensure that appropriate safeguards are in place to prevent
unintended lasing.
The use of the Quick Start Plug bypasses the laser’s safety interlock
function, potentially exposing personnel in the area to invisible
infrared laser radiation.
The Quick Start Plug is intended only for initial testing and troubleshooting by qualied personnel. In normal operation, the laser’s
Remote Interlock input should be connected to the machine’s
safety interlock circuitry.
SYNRAD f201 Operator’s Manual Version 3.2
45
Page 48
operation
Controls and indicators
The Controls and indicators section includes subsections:
■ f201 lasers
f201 lasers
2
3
1
AVOID EXPOSURE
Invisible laser radiation
is emitted from
this aperture.
Figure 3-1 F201 controls and indicators (front).
1
Laser Aperture – provides an opening in Firestar’s front panel from which the beam exits.
2
Aperture Seal – prevents dust from damaging the output coupler during shipping. Remove the
red self-adhesive label before applying power to the laser.
3
Optical Accessories Mounting – provides six threaded holes (8–32 UNC) for mounting optional
beam delivery components available from SYNRAD. Because excessive weight may damage the
laser, consult SYNRAD before mounting components not specically designed as Firestar options. Refer to Firestar package outline drawings in the Technical Reference chapter for mounting
dimensions.
Note: When mounting optical components to f201 lasers, the 8–32 UNC fasteners must extend no
further than 4.8 mm (0.19") into the laser’s faceplate.
SYNRAD f201 Operator’s Manual Version 3.2
46
Page 49
operation
10
9
8
76
5
FUSE
Remote
Interlock
LaseShutterReadyTemp
5
WATER OUT
USER I/O
GAS PURGE
Clean and dry
air or N2 only
11
DC POWER
20A FAST
96 VDC 36A MAX
4
12
WATER IN
Figure 3-2 F201 controls and indicators (rear).
4
WATER OUT and WATER IN Ports – provides connection points for Firestar’s cooling system using
non-removable straight 1/2-inch ttings. A cooling kit contains 90° tubing adaptors for both 1/2inch and 12-mm cooling tubing.
5
FUSE Holder – protects laser circuitry with a 20 A, 125 V fast-acting fuse.
6
DC POWER Connector – receives 96 VDC from the DC power supply through the DC Power cables.
7
Remote Interlock Indicator – illuminates green to indicate that a remote interlock circuit is closed
and that lasing may be enabled. The Remote Interlock indicator is red and lasing is disabled if the
interlock input is open.
8
Temp Indicator – illuminates green to indicate that coolant temperature and ow is within limits
and that lasing may be enabled. The Temp indicator is red and lasing is disabled if the coolant temperature rises above safe operating limits.
9
Ready Indicator – illuminates yellow when the laser is enabled, indicating that, after a ve-second
delay, lasing will begin when a PWM Command signal is applied.
10
Lase Indicator – illuminates red to indicate that the laser is actively lasing. The Lase indicator is o
when tickle pulses are being generated and illuminates red when PWM Command signal pulses
are long enough to produce laser output.
11
User I/O Connector – provides a connection point for auxiliary output power, as well as input and
output signals. Refer to User I/O connections in the Technical Reference chapter for pinouts and
signal descriptions.
12
Gas Purge Connector (f201 only) – provides a low pressure nitrogen (or pure air) purge gas con-
nection to prevent dust and debris from entering the laser housing and damaging electronic or
optical components.
SYNRAD f201 Operator’s Manual Version 3.2
47
Page 50
operation
Initial start-up
The Initial start-up section includes subsections:
■ With a UC-2000 Controller
■ Without a UC-2000 Controller
With a UC-2000 Controller
Before your laser is put into service for the rst time, its functionality should be veried. Follow this
procedure to verify the laser system is operating at optimum performance. For this procedure, use the
UC-2000 as a stand-alone controller; do not attempt to control the laser or UC-2000 externally.
Note: When performing the initial start-up sequence, you must rst connect the Quick Start Plug or you
must provide the required Remote Interlock and Shutter Open Request signals to the User I/O
connector. See User I/O connections in the Technical Reference chapter for User I/O pinouts and
signal descriptions.
Starting auxiliary equipment
1
Ensure that all personnel in the area are wearing protective eyewear.
2
Remove the red self-adhesive aperture seal from the laser faceplate.
3
Place a beam block 61 cm (24 in) from the laser aperture to prevent the beam from traveling beyond
the work area.
4
Turn on the chiller and set the temperature setpoint between 18 °C–22 °C. Verify that the chiller is
delivering the recommended ow rate for your laser at less than 414 kPa (60 PSI) of pressure. Examine all cooling connections carefully and ensure that they do not leak.
5
Start purge gas ow at a rate of 0.85–1.7 m3/hr (30–60 SCFH, Standard Cubic Feet per Hour) at a
pressure not to exceed 34.5 kPa (5 PSI). If a owmeter is not available, set a purge pressure between
13.8–34.5 kPa (2–5 PSI).
Note: If you have not yet operated your UC-2000 Universal Laser Controller, refer to the UC-2000 Laser
Controller Operator’s Manual and quick start guide for setup and operation instructions before
continuing.
6
Set the UC-2000 to MANUAL mode, and then set the PWM Adj Knob to provide zero percent
output (0.0%). The UC-2000’s Lase indicator should be O.
7
Verify that the laser’s Keyswitch (if equipped) is in the OFF/RESET position.
8
Turn on the +96 VDC power supply.
If the factory-wired Quick Start Plug is installed, the Remote Interlock indicator will illuminate green
and the Shutter indicator will illuminate blue—if the manual Shutter Switch is Open. The Temp indicator will illuminate green if laser temperature is within safe operating limits.
SYNRAD f201 Operator’s Manual Version 3.2
48
Page 51
operation
Starting your f201 laser
Note: Ready LEDs denote separate control functions. Although the Ready lamp may light, no power is
applied to the RF boards until both Ready and Shutter indicators are illuminated.
1
If the laser has a Diode Pointer installed, remove its aperture dust cover.
Note: Assure the red self-adhesive aperture seal is removed from the laser faceplate.
2
Press the UC-2000’s Lase On/O button. The Lase indicator on the UC-2000 should illuminate.
3
Use the PWM Adj Knob on the UC-2000 to slowly increase power. The Lase LED turns red when
PWM Command pulses are long enough to produce laser output (typically 7–10 µs at 5 kHz). The
spot where the beam hits the beam block increases in brightness to indicate increased power
output.
4
Press the UC-2000’s Lase On/O button to stop lasing. Lase indicators on the UC-2000 and the
laser should both turn o.
5
Shut o the chiller or otherwise stop coolant ow through the laser.
6
Shut o gas purge ow to the laser. In dirty or dusty environments; however, it may be necessary
to purge the laser continuously to prevent contamination of internal optics.
Warning
serious
personal
Because of phase dierences, external tickle pulses may combine
with the internally-generated tickle signal causing the Lase LED to
icker during the transition from tickle to lasing. Laser output may
occur if the Lase LED ickers.
injury
If your f201 laser fails to lase, refer to Troubleshooting in the Maintenance/Troubleshooting chapter for
troubleshooting information.
Without a UC-2000 Controller
If you are not using a UC-2000 to control the laser, follow the procedure below to verify laser operation.
Although a tickle signal is not required, you will need to provide PWM Command signals to User I/O
connector. Refer to User I/O connections in the Technical Reference chapter for connector pinouts and
refer to Controlling laser power in the Technical Reference chapter for Command signal descriptions.
Note: When performing the initial start-up sequence, you must rst connect the Quick Start Plug or
you must provide the required Remote Interlock Request signals to the User I/O connector. See
User I/O connections in the Technical Reference chapter for User I/O pinouts and signal descriptions.
Caution
possible
equipment
Do not ow coolant through the laser for an extended period of
time when the laser is shutdown. This causes condensation to form
inside the laser which may result in catastrophic damage to internal
optics and electronic circuits.
damage
SYNRAD f201 Operator’s Manual Version 3.2
49
Page 52
operation
Initial start-up
Starting auxiliary equipment
1
Ensure that all personnel in the area are wearing protective eyewear.
2
Remove the red self-adhesive aperture seal from the laser faceplate.
3
Place a beam block 61 cm (24 in) from the laser aperture to prevent the beam from traveling beyond
the work area.
4
Connect the output of your PWM controller to PWM Input (Pin 9) on the laser’s User I/O connector
and connect the ground or return of the controller to PWM Return (Pin 1). On f200 lasers, parallel
this PWM signal between both User I/O connectors.
5
Turn on the chiller and set the temperature setpoint between 18 °C–22 °C. Verify that the chiller is
delivering the recommended ow rate for your laser at less than 414 kPa (60 PSI) of pressure. Examine all cooling connections carefully and ensure that they do not leak.
6
Start purge gas ow at a rate of 0.85–1.7 m3/hr (30–60 SCFH, Standard Cubic Feet per Hour) at a
pressure not to exceed 34.5 kPa (5 PSI). If a owmeter is not available, set a purge pressure between 13.8–34.5 kPa (2–5 PSI).
Caution
possible
Inlet cooling water temperature must always be maintained above the
dew point to prevent condensation and water damage to your Firestar
laser.
equipment
damage
7
Verify that the laser’s Keyswitch (if equipped) is in the OFF/RESET position.
8
Turn on the +96 VDC power supply.
Note: If the factory-wired Quick Start Plug is installed, the Remote Interlock indicator will illuminate
green and the Shutter indicator will illuminate blue—if the manual Shutter Switch is open. The
Temp indicator will illuminate green if laser temperature is within safe operating limits.
Caution
possible
equipment
damage
Do not exceed a gas purge pressure of 34.5 kPa (5 PSI). Excessive
pressure may damage the purge assembly or other internal laser
components.
Do not use argon as a purge gas. Use only nitrogen or clean, dry air
as described in the following table, Purge gas specications.
SYNRAD f201 Operator’s Manual Version 3.2
50
Page 53
operation
This page intentionally left blank.
SYNRAD f201 Operator’s Manual Version 3.2
51
Page 54
technical reference
4
Use information in this chapter as a technical reference for your Firestar f201 laser.
■ Packaging instructions – illustrates how to package Firestar f201 lasers for ship-
ment.
SYNRAD f201 Operator’s Manual Version 3.2
52
Page 55
technical reference
Technical overview
The Technical overview section includes subsections:
■ Laser tube
■ Optical resonator
■ Internal RF power supply
■ Optical setup
Laser tube
Firestar® f201 lasers were developed using new technology patented by SYNRAD. Firestar’s patented “f”
technology, based on an unstable free-space resonator design, enables SYNRAD to economically produce a symmetrical laser beam from a small but powerful laser capable of operating for many years with
virtually no maintenance. Firestar’s unique extruded aluminum envelope oers excellent heat transfer,
long gas life, and low operating costs in contrast to other laser tube technologies. Besides being the
vessel that maintains the lasing environment, the aluminum tube is also the structural platform that
integrates the laser’s optical, electrical, and cooling components.
Duo-Lase
Firestar lasers consist of two standard, sealed tubes (f201 tubes respectively) that are combined using
SYNRAD’s Duo-Lase® technology to provide a single diraction-limited beam at twice the output power.
All power and control functions between the two laser sections are totally independent, essentially
achieving fail-safe operation for applications that can be served with the power of a single laser tube.
Any one electronic or laser tube failure will aect only that section, leaving the second section unaffected and available for temporary use.
The Duo-Lase optical combining technique is based on the fact that each tube is linearly polarized,
which allows the use of a polarization-sensitive beam combiner to achieve a 98 percent eciency in
combining the two beams. Both components of the resulting output beam are spatially parallel and
collinear. The normal temporal and spatial variations inherent in a single tube laser are reduced by combining the output of two lasers.
®
operation
Cooling
Heat generated by excited CO2 molecules is transferred to the bore walls by diusion. Collected heat is
transferred to the water in the cooling tubes by conduction of the electrodes and aluminum envelope.
The coolant path is directed through corrosion-resistant copper alloy tubing to regulate laser temperature for maximum stability.
Optical resonator
The optical resonator, in conjunction with the electrodes and the gas mixture, generates the laser beam.
Firestar f201 optical resonators are comprised of three optical elements: a front mirror, a rear mirror, and
an output window. These optical elements are fastened to the tube’s exterior and are exposed to its
interior through holes in the end caps. O-rings are sandwiched between optical elements and the end
cap to form a gas seal and to provide a exible cushion that allows the slight movement necessary for
alignment. All optical elements are aligned and locked into place by factory technicians before the laser
is shipped.
SYNRAD f201 Operator’s Manual Version 3.2
53
Page 56
technical reference
Full Angle Divergence 4 mrad
Technical overview
Caution
possible
equipment
damage
On f201 lasers, a small telescope just before the output coupler corrects a slight ellipticity in the beam
generated by the resonator. The output beam, roughly square as it exits the telescope, transitions to
Gaussian-like in mid and far elds three meters and beyond. The internal structure and optics of the
resonator combine to produce a Gaussian-like mode quality (M2 factor) of 1.2 ±0.1. As shown in the gure below, beam diameter (beam waist) is 3.5 mm at the output aperture and full angle divergence due
to diraction is 4 milliradians (a 4 mrad full angle divergence means that beam diameter increases 4 mm
over every one meter distance traveled).
Because of their smaller beam diameter, Firestar lasers have signicantly higher power densities than previous SYNRAD lasers. This
means that any contamination on the laser’s output coupler (or on
any beam delivery optic) can absorb enough energy to damage one or
more optics in the beam path. Periodically inspect the laser’s output
coupler and all other beam delivery optics for signs of contaminants
and then carefully clean as required. In dirty environments, purge
laser optics using ltered air or nitrogen to prevent vapor and debris
from accumulating on optical surfaces.
Ø
Beam Diameter 3.5 mm
at Output Aperture
Figure 4-1 beam characteristics.
On f201 lasers, the beam waist is located one meter before the output aperture due to the addition of
internal beam conditioning optics that improve beam quality and strip any side lobes present to produce a Gaussian-like mode quality (M2 factor) of 1.2 ±0.1. Beam diameter at the output aperture is 4.5
mm with a full angle divergence of 4 milliradians.
Internal RF power supply
The compact, self-contained RF power supply converts 96 VDC input power into a high-power radio
frequency (RF) signal. The 40.68 MHz RF output from the RF power supply is then routed to the electrode
structure in the laser tube where it excites carbon dioxide (CO2) gas in the tube to produce lasing.
Control circuit
Switches and sensors built into the f201 control board monitor the laser for various conditions like
under/over voltage, over temperature, and VSWR faults that pose a risk of damage to the laser. Additionally, laser operation is controlled by the following: a manual Shutter Switch; a Keyswitch (on Keyswitchequipped lasers); the Shutter Open Request input; the Remote Interlock input; and the Remote Reset/
Start Request input.
SYNRAD f201 Operator’s Manual Version 3.2
54
Page 57
technical reference
Technical overview
Optical setup
After selecting a laser for a CO2 laser processing system, the two most important elements to consider are:
(1) beam delivery optics to transmit the beam to the work area; and (2) focusing optics to focus the beam
onto the part or material to be processed. Each element is crucial in the development of a laser-based material processing system and each element should be approached with the same careful attention to detail.
Delivery optics
Divergence, or expansion, of the laser beam is important in materials processing since a larger beam
entering the focusing optic produces a smaller focused spot. Because the laser beam diverges slowly,
increasing by two or three diameters over several meters, Firestar lasers should be mounted a distance
of 1.3–1.8 meters (50–70 inches) away from the work area and no closer than one meter (40 inches) for
optimum performance. Right angle turning mirrors are often used in conjunction with the laser mounting position to obtain this distance. Figure below shows how right angle turning mirrors in a “ying
optics” setup help create this longer beam path.
Firestar f201 lasers, with their internal one-meter beam path and conditioning optics, can be mounted
right at the work area.
Figure 4-2 “Flying optics” beam path.
Note: Optical components in the beam path must always be aligned to the actual beam path, not the
laser faceplate. Because of slight variations in laser construction, the beam path may not always
be centered in, or perpendicular to, the aperture in the faceplate.
Expander/collimators are optical devices that reduce beam divergence while at the same time increasing
beam diameter by a selectable magnication factor. Adding an expander/collimator to the “ying optics”
setup shown above would substantially reduce beam divergence and any variance in beam diameter caused
by the changing optical path length. In xed-length delivery systems where the laser is positioned only one
meter away from the focusing optic and a small spot size is required, an expander/collimator is again the
best solution to provide the required beam expansion before reaching the focusing optic.
SYNRAD f201 Operator’s Manual Version 3.2
55
Page 58
technical reference
Technical overview
Focusing optics
When selecting a focusing optic, the primary consideration should be material thickness and any vertical
tolerances that occur during nal part positioning rather than making a selection based only on minimum
spot size. The chosen focal length should create the smallest possible focused spot while providing the
depth of eld required for the material to be processed.
Caution
possible
equipment
damage
Optics are fragile and must be handled carefully, preferably by the mounting ring only. Be careful to select
optics that are thick enough to withstand the maximum assist gas pressure available for the process. This
is especially important in metal cutting applications using high-pressure assist gases. Cleanliness is another
important issue aecting performance and becomes increasingly important as laser power increases. Dirty
or scratched lenses will under perform, exhibit a vastly shortened lifetime, and may fail catastrophically.
When the laser application requires air as an assist gas, use only breathing quality air available in cylinders from a welding supply company. Compressed shop air contains minute particles of oil and other
contaminants that will damage optical surfaces. If compressed shop air is the only choice available, it
must be ltered to the specications shown in the following table.
Trace amounts of contamination on the laser’s output window (or
on any optic in the beam path) can absorb enough energy to damage
the optic. Inspect all beam delivery optics periodically for signs of
contaminants and carefully clean as required. In dirty environments,
purge laser optics using ltered air or nitrogen to prevent vapor and
debris from accumulating on optical surfaces.
Table 4-1 Assist gas purity specications.
Assist Gas Typical Purpose Specication
Air Cutting/Drilling Breathing Grade > 99.9996% purity; ltered to ISO Class 1
particulate level
Air Cutting/Drilling Compressed Instrument-grade air ltered and dried to
ISO 8573-1:2010 Class 1, 2, 1 (< 10 1.0–
5.0 µm particles/m3; < –40 °F dew point;
< 0.01 mg/m3 oil vapor)
Argon Welding High Purity Grade > 99.998% purity; ltered to ISO Class 1 particulate level
Helium Welding High Purity Grade > 99.997% purity; ltered to ISO Class 1 particulate level
Nitrogen Cutting/Drilling High Purity Grade > 99.9500% purity; ltered to ISO Class 1
particulate level
Oxygen Cutting/Drilling Ultra Pure Grade > 99.9998% purity; ltered to ISO Class 1
particulate level
SYNRAD f201 Operator’s Manual Version 3.2
56
Page 59
technical reference
Controlling laser power
The Controlling laser power section includes subsections:
■ Control signals
■ Operating modes
Control signals
Much of the information provided in this section describes the use of a SYNRAD UC-2000 Universal Laser Controller to provide PWM Command signals to the f201 laser. If using an alternate method of laser
control, thoroughly review this section, Controlling laser power, as well as the following section, User
I/O connections, for an understanding of the signal requirements necessary to control Firestar lasers. For
more information about the UC-2000, please consult the UC-2000 Laser Controller Operator’s Manual.
Tickle pulse
Tickle pulses pre-ionize the laser gas to just below the lasing threshold so that a further increase in pulse
width adds enough energy to the plasma to cause laser emission. Tickle pulses cause the laser to respond predictably and almost instantaneously to PWM Command signals, even when there is considerable delay (laser o time) between applied Command signals. All Firestar f201 lasers incorporate a builtin tickle generator, freeing customers from the need to supply tickle pulses between lasing commands.
Internal circuitry monitors the incoming PWM signal and determines the amount of time the laser was
on (lasing) during the last 200 microsecond (µs) interval. If the laser’s on time was greater than the preset
tickle value, then no tickle pulse is generated because the PWM signal was sucient to maintain a plasma
state. If no PWM signal was applied during the 200-µs measurement period (or was shorter than the preset
tickle value), internal circuitry generates a tickle pulse such that the laser always receives a pre-set amount
of RF drive averaged over any 200-µs interval.
Warning
serious
personal
Because of phase dierences, external tickle pulses may combine with
the internally-generated tickle signal causing the Lase LED to icker
during the transition from tickle to lasing. Laser output may occur if
the Lase LED ickers.
injury
Pulse Width Modulation (PWM)
Pulse Width Modulation, or PWM, controls laser power by varying the duty cycle of Firestar’s RF ampliers, which in turn control the time-averaged RF power applied to the laser. Typically, laser output follows
the PWM input with a rise and fall time constant of ~150 µs; however, the laser cannot precisely follow
PWM input signals if the “On” pulse is less than 150 µs in duration. At a constant 50% duty cycle, f201
lasers typically reach 90–100% of full optical output when operated at a frequency of 3 kHz.
The percentage of optical output increases as duty cycle increases (at a constant PWM frequency) or as
PWM frequency decreases (at a constant duty cycle). Figure on the following page shows representative
f201 optical output waveforms at two dierent duty cycles and frequencies.
SYNRAD f201 Operator’s Manual Version 3.2
57
Page 60
technical reference
22
1
Optical output pulse (10% duty cycle at 100 Hz)
Optical output pulse (50% duty cycle at 5 kHz)
Controlling laser power
1
T
T
T
100mV Ch2 5 V M 250µs Ch2 1.6 V
Ch1
22
Ch1
T
50mV Ch2 5 V M 100µs Ch2 1 V
Figure 4-3 Typical Firestar f201 waveforms.
Firestar f201 lasers are designed to operate at Command signal base frequencies up to 100 kHz; however, the choice of PWM frequency depends on the user’s specic application. In the majority of laser
applications, the UC-2000’s default Command signal frequency of 5 kHz has proven to work well. When
considering Command frequencies at 5 kHz or below, please review Marking/engraving operation later
in this section. For high-speed motion applications that cannot tolerate any ripple in the optical beam
response but still need adjustable power levels, we recommend the use of higher PWM frequencies, up
to 100 kHz maximum. At 100 kHz, the laser’s optical beam response no longer follows the Command
input and is very nearly a DC value with just a small amount of ripple present.
Command signal
The modulated Command signal applied between Pin 9, PWM Input, and Pin 1, PWM Return, of the User
Warning
Always use shielded cable when connecting your PWM Command
signal source to PWM Input/PWM Return inputs. In electrically-
serious
personal
noisy environments, long lengths of unshielded wire act like an
antenna and may generate enough voltage to trigger uncommanded
lasing.
injury
I/O connector on the Firestar f201 laser has three basic parameters: signal amplitude, base frequency,
and PWM duty cycle. By changing these parameters, you can command the beam to perform a variety
of marking, cutting, welding, or drilling operations.
The rst Command signal parameter, signal amplitude, is either logic low—corresponding to laser beam
o, or logic high—corresponding to beam on. The laser o voltage, typically 0 V, can range from 0.0 V to
+0.8 VDC while the laser on voltage, typically 5 V, can range from +3.5 V to +6.7 VDC.
Base frequency, the second parameter, is the repetition rate of the PWM input signal. The standard base
frequency is 5 kHz, which has a period of 200 µs. Maximum PWM frequency is 100 kHz.
SYNRAD f201 Operator’s Manual Version 3.2
58
Page 61
technical reference
0
5
200 µs
200 µs
Controlling laser power
The third Command signal parameter, PWM duty cycle, is the percentage of the period that the Command signal is high. If the Command signal’s amplitude (at 5 kHz) is high for 100 µs and low for 100 µs, it
has a 50% duty cycle; if the amplitude is high for 190 µs and low for 10 µs, it has a 95% duty cycle. Figure
below illustrates PWM Command signal parameters while the table below lists PWM signal specications.
100 µs
VDC
VDC
5 kHz Command Signal at 50% Duty Cycle
190 µs
5 kHz Command Signal at 95% Duty Cycle
Figure 4-4 PWM Command signal waveform.
Firestar’s User I/O PWM input consists of a high-speed optoisolator LED with a forward voltage drop (Vf)
of 1.5 VDC. The PWM input frequency can range from DC (0 Hz) to 100 kHz. Table below provides minimum, maximum, and nominal PWM signal specications.
In addition to controlling the Firestar laser using a UC-2000 Controller, control of Firestar externally,
without a UC-2000, is also possible. The two primary elements of laser control are gating, the ability to
turn the laser on and o at the appropriate times, and power, the ability to control the laser’s output energy. Both gating and power can be handled by a device such as a personal computer, Programmable Logic
Controller (PLC), or function generator capable of sending PWM pulses at the proper time (gating) and
with the proper duty cycle (power).
Analog voltage or current control
Although Firestar f201 lasers cannot be controlled directly by analog voltage or current signals, this type
of control is possible when using the UC-2000 Controller. The Controller is connected normally to the
laser and analog voltage or current signals sent to the UC-2000’s ANV/C connector then control both
laser gating and power.
SYNRAD f201 Operator’s Manual Version 3.2
59
Page 62
technical reference
Controlling laser power
To generate the correct analog voltage from a computer or PLC, a Digital-to-Analog (D/A or DAC) card
capable of generating 0 V (laser o) to 10 V (maximum laser power) must be installed. To generate the
proper analog current, install a D/A card that can generate 4 mA (laser o) to 20 mA (maximum power).
Software able to control your analog output card is required for either conguration.
Continuous wave (CW)
In some applications, such as high speed marking or cutting, the time constant of the laser and the PWM
modulation causes a series of dots that may be visible on the marking surface instead of a “clean” line.
Operating the laser in CW mode will prevent this behavior from occurring.
To operate the laser in CW mode, apply a constant +5 VDC signal to Pin 9, PWM Input, and Pin 1, PWM
Return, on the User I/O connector. This constant voltage source forces the internal switching electronics
to remain on, providing continuous and uninterrupted laser output power. During CW operation, output
power cannot be changed. To adjust output power, refer back to the Pulse Width Modulation (PWM) section for information regarding high frequency operation.
Note: SYNRAD lasers are designed for maximum performance using a 95% duty cycle. Increasing the
maximum PWM percentage beyond 95% greatly increases the laser’s heat load with little or no
corresponding increase in laser output power. Continuous operation at 99% duty cycle may lead
to thermal instability and optical degradation.
Gated operation
In many marking and cutting applications, the laser is required to pulse, or gate, on and o in synchronization with an external control signal (typically from a computer or function generator operating in the
range from DC to 1 kHz). To pulse or gate the laser, connect a signal providing +5.0 VDC pulses to the
Gate connector on the rear panel of the UC-2000.
Users who intend to use a gating signal should set the UC-2000’s gate input logic to internal Pull-Down
(normally o) mode. This prevents the beam from being enabled unless a high level (+3.5 V to +5.0
VDC) signal is applied to the Gate input connector. In the pull-down (normally o) mode an asserted
logic low state, short circuit to ground, or an open or disconnected Gate input locks the beam o.
Warning
serious
personal
injury
The UC-2000’s default gate logic is factory set to internal Pull-Up
(normally on) mode so that an open (disconnected) Gate input
causes the laser to turn on. This functionality allows the user to easily
test and verify laser operation prior to integration.
In an integrated system, you should congure the UC-2000’s gate
input logic to internal Pull-Down (normally o) mode. This prevents
the beam from being enabled unless a high level (+3.5 V to +5.0
VDC) signal is applied to the Gate input connector. In the PullDown (normally o) mode an asserted logic low signal, short circuit
to ground, or an open or disconnected Gate input locks the beam o.
Many CO2 lasers operating in applications requiring short gating pulses at repetition rates below 500 Hz
will exhibit some leading edge overshoot regardless of the PWM frequency. This occurs because a cooler
lasing medium (the CO2 gas) is more ecient than a hotter one. This overshoot eect is more pronounced
at lower gating frequencies since the gas has a longer time to cool down between Command signal pulses.
SYNRAD f201 Operator’s Manual Version 3.2
60
Page 63
technical reference
Controlling laser power
Marking/engraving operation
When the delay between the end of one PWM Command signal pulse and the beginning of the next
PWM pulse exceeds 200 microseconds (less than or equal to 5 kHz), Firestar’s on-board tickle generator
sends a tickle pulse to maintain plasma ionization in the tube. Because the on-board tickle generator can
not anticipate when the next PWM Command pulse will arrive; the tickle pulse (which typically lasts for
2–6 µs depending on the laser) can eectively merge with a PWM signal that follows closely afterwards.
When the PWM pulse that follows is short, causing the tickle pulse to become a signicant fraction of the
PWM pulse duration, then the tickle pulse eectively substantially increases the length of the PWM pulse
it has merged with. For subtle marking applications on sensitive, low threshold materials this lengthened
PWM pulse may aect mark quality.
While this situation can occur when using PWM Command signal frequencies of 5 kHz and less, it is important to note that it isn’t the Command signal frequency itself that is the determining factor but rather
this behavior happens only when the o time between PWM pulses exceeds 200 microseconds.
SYNRAD f201 Operator’s Manual Version 3.2
61
Page 64
technical reference
User I/O connections
The User I/O connections section includes subsections:
■ User I/O connection summary
■ Input/output signals
■ Sample I/O circuits
■ I/O connections
User I/O connection summary
Table below provides a quick reference summary to Firestar f201 User I/O connections.
Table 4-3 User I/O pin descriptions.
Pin Function Description
1 PWM Return
Use this input pin as the return side of the PWM Command signal.
2 Remote Reset/Start Request input
Apply a positive or negative voltage (±5–24 VDC) with respect to Pin 11, Input Common,
to reset or remote keyswitch the laser. The laser remains disabled while voltage is applied.
Removing voltage from the Remote Reset/Start Request input causes the laser’s Ready
indicator to illuminate and begins a ve-second countdown after which lasing is enabled.
3 Remote Interlock input
Apply a positive or negative voltage (±5–24 VDC) with respect to Pin 11, Input Common,
to enable lasing. If your system does not use a remote interlock, this input must be connected to a voltage source in the range of ±5–24 VDC. Refer to Figure 3-7 for a diagram
showing how the Remote Interlock input is factory-jumpered on the Quick Start Plug.
4 + 5 VDC Auxiliary Power
This connection provides +5 VDC for driving external inputs or outputs. The +5 VDC
Auxiliary Power output can source up to 0.5 A and is protected by a 0.5 A self-resetting
fuse. The return (ground) path must be through Pin 12, Auxiliary DC Power Ground.
5 + 24 VDC Auxiliary Power
This connection provides +24 VDC for driving external inputs or outputs. The +24 VDC
Auxiliary Power output can source up to 0.5 A and is protected by a 0.5 A self-resetting
fuse. The return (ground) path must be through Pin 12, Auxiliary DC Power Ground.
6 Laser Active output
This bi-directional switched output is internally connected to Pin 13, Output Common,
when the laser is actively lasing (Lase indicator illuminated red). This output is open (high
impedance) when no beam is being emitted (Lase indicator O).
7 Over Temperature output
This bi-directional switched output is internally connected to Pin 13, Output Common,
when laser temperature is above safe operating limits (Temp LED illuminated red). The
output is open (high impedance) when laser temperature is within limits (Temp LED green).
SYNRAD f201 Operator’s Manual Version 3.2
62
Page 65
technical reference
User I/O connections
Table 4-3 User I/O pin descriptions (continued).
Pin Function Description
8 Laser Ready output
This bi-directional switched output is internally connected to Pin 13, Output Common,
when the laser is enabled (Ready LED illuminated yellow), indicating that lasing will occur
when a PWM Command signal is applied to Pin 9 and Pin 1. When this output is initially
switched closed, there is a ve-second delay during which lasing is inhibited. This output is
open (high impedance) when the laser is disabled (Ready LED O).
9 PWM Input
Connect your PWM Command signal (+5 VDC, 5 kHz nominal, 100 kHz max, pulse width
modulated) to this input pin to control laser output power. Refer back to Controlling laser
power for further information on laser control signals.
11 Input Common
Use this input pin to connect return lines for Remote Interlock, Shutter Open Request,
and Remote Reset/Start Request lines.
12 Auxiliary DC Power Ground
This connection provides a ground (earth) connection for +5 and +24 VDC auxiliary
power outputs. This pin is the only User I/O pin that is connected to chassis ground. Do not
use this pin for grounding if DC power to external I/O circuits is supplied from an external
customer-supplied DC power source.
13 Output Common
Use this pin to complete the return path for output connections (Pin 6, 7, 8, 14, or 15).
The Output Common line is protected by a 0.25 A self-resetting fuse.
15 Interlock Open output
This bi-directional switched output is internally connected to Pin 13, Output Common,
when remote interlock circuitry is open (Remote Interlock indicator illuminated red), indicating that lasing is disabled. The output is open (high impedance) when lasing is enabled
(Remote Interlock indicator green).
Input/output signals
Firestar’s input/output signals are divided into three categories: auxiliary DC power, input signals, and
output signals. Signals in each category are fully described in the sections below. Following gure illustrates the pin arrangement of the User I/O (15 pin female D-type subminiature) connector on Firestar’s
rear panel.
SYNRAD f201 Operator’s Manual Version 3.2
63
Page 66
technical reference
User I/O connections
Auxiliary DC power
Firestar’s User I/O connector provides auxiliary DC power for driving external inputs or outputs connected
to the User I/O port. Pin 4, +5 VDC Auxiliary Power, and Pin 5, +24 VDC Auxiliary Power, are protected
by self-resetting fuses rated at 0.5 A. Pin 12, Auxiliary DC Power Ground, is connected to chassis ground
while all other User I/O pins are oating with respect to chassis ground. Figure 3-6 illustrates Firestar’s
internal DC supply wiring.
Caution
Turn o DC power before installing or removing any plug or cable
from the User I/O connector. Ensure that user connections are made
possible
equipment
to the appropriate pins and that the appropriate signal levels are applied. Failure to do so may damage the laser.
damage
Pin 8
FUSE
20A FAST
DC POWER
96 VDC 18A MAX
Figure 4-5 User I/O connector pinouts.
Remote
Interlock
USER I/O
Pin 1
OFF/RESET
LaseShutterReadyTemp
ON
Pin 9Pin 15
Pin 4 + 5 VDC Auxiliary Power
This connection provides +5 VDC for driving external inputs or outputs. The +5 VDC Auxiliary
Power output can source up to 0.5 A and is protected by a 0.5 A self-resetting fuse. The return
(ground) path must be through Pin 12, Auxiliary DC Power Ground.
Pin 5 + 24 VDC Auxiliary Power
This connection provides +24 VDC for driving external inputs or outputs. The +24 VDC Auxiliary Power output can source up to 0.5 A and is protected by a 0.5 A self-resetting fuse. The
return (ground) path must be through Pin 12, Auxiliary DC Power Ground.
Pin 12 Auxiliary DC Power Ground
SYNRAD f201 Operator’s Manual Version 3.2
64
Page 67
technical reference
This connection provides a ground (earth) connection for +5 and +24 VDC auxiliary power outputs. This pin is the only User I/O pin that is connected to the laser’s chassis ground. Do not use
this pin for grounding if I/O circuits are powered from an external customer-supplied DC power
source.
Figure 4-6 Auxiliary power supply wiring.
Input signals
A total of four user inputs allow control of Firestar lasers. Remote Interlock, Shutter Open Request, and
Remote Reset/Start Request inputs are optoisolated and bi-directional, allowing for positive or negative
polarity signal inputs. These three signals also share a common return connection, Input Common, which
is separate from chassis ground to completely isolate control signals for optimal EMI performance. The
fourth input, PWM Input, is optoisolated and has a separate return, PWM Return, to fully isolate PWM
signals from the other three user inputs. Note that throughout this manual, input voltage levels are specied with respect to their corresponding return line.
Pin 1 PWM Return
Connect the return side of your PWM Command signal to this pin. Refer to Table 4-4 for input
circuit specications.
Pin 2 Remote Reset/Start Request
Apply a positive or negative voltage (±5–24 VDC) with respect to Pin 11, Input Common, to reset or remote keyswitch the laser. The laser remains disabled while voltage is applied to this input.
Removing voltage from the Remote Reset/Start Request input causes the laser’s Ready lamp
to illuminate and begins a ve-second countdown after which lasing is enabled. Because all DC
power is removed from the laser’s RF driver when this input is active, no lasing can occur until
voltage is removed from Pin 2. Refer to Table 3-4 for input circuit specications.
For Keyswitch-equipped lasers in automated systems, the Remote Reset/Start Request input
can operate as a remote keyswitch. To use this “remote keyswitch” functionality, rst place the
Keyswitch in the ON position. Then after each DC power-up cycle (or to reset a fault condition),
apply a momentary voltage pulse in the range of ±5–24 VDC to Pin 2. This reset action initiates
a ve-second delay after which lasing is enabled. To prevent operator injury when using Keyswitch-equipped lasers, the Remote Reset/Start Request signal is ignored when the Keyswitch is
set to the OFF/RESET position. The manually-operated Keyswitch must be set to the ON position
before the laser can be reset from a remote location.
SYNRAD f201 Operator’s Manual Version 3.2
65
Page 68
technical reference
User I/O connections
Pin 3 Remote Interlock
Apply a positive or negative voltage (±5–24 VDC) with respect to Pin 11, Input Common, to
enable lasing. If your system does not use a remote interlock, this input must be connected to a
voltage source in the range of ±5–24 VDC. Refer to following gure for a diagram showing how
the Remote Interlock input is factory-jumpered on the Quick Start Plug. Because all DC power is
removed from the laser’s RF driver when this input is inactive, no lasing can occur until voltage
is applied to Pin 3. Refer to Table 4-4 for input circuit specications.
After voltage is removed and then re-applied to the Remote Interlock input on Keyswitchequipped lasers, the Keyswitch or the Remote Reset/Start Request input must be toggled to
restart the laser This action initiates a ve-second delay during which lasing is inhibited.
On OEM lasers, the Remote Interlock input is not latched. Re-applying a voltage to Pin 3
enables the Ready indicator and lasing is possible after the ve-second delay, provided that the
Shutter Open Request signal is also present (Shutter indicator illuminated).
Pin 9 PWM Input
Connect your PWM Command signal (+5 VDC, 5 kHz nominal, 100 kHz max) to Pin 9. This pulse
width modulated Command signal controls laser output so that a duty cycle of 50% corresponds to a laser output of approximately one-half rated output power and a duty cycle of
95% corresponds to approximately full output power. Refer to Controlling laser power in this
chapter for further information on laser control signals. Connect the PWM signal source return
to Pin 1, PWM Return. See following table for input circuit specications.
Pin 10 Shutter Open Request
Apply a positive or negative voltage (±5–24 VDC) with respect to Pin 11, Input Common, to
enable lasing, in conjunction with the manual Shutter Switch (if equipped). If your system does
not supply a Shutter Open Request signal, this input must be connected to a voltage source
in the range of ±5–24 VDC. Refer to Figure 3-7 for a diagram showing how the Shutter Open
Request input is factory-jumpered on the Quick Start Plug. When a voltage is initially applied to
this input, there is a ve-second delay during which lasing is inhibited. Because all DC power is
removed from the laser’s RF driver when this input is inactive, no lasing can occur until voltage is
applied to Pin 10. See Table 3-4 for input circuit specications.
To enable Keyswitch-equipped lasers, you must apply a voltage (in the range of ±5–24 VDC) to
the Shutter Open Request input and move the manual Shutter Switch (if equipped) to the Open
position. Lasing is inhibited when voltage is removed from Pin 10 or the Shutter Switch is Closed.
Note: To ensure optimum performance, a ve-second delay is imposed from the time a Shutter
Open Request input occurs to the time that PWM Command signals are accepted.
Pin 11 Input Common
Use this pin to connect return lines for Remote Interlock, Shutter Open Request, and Remote
Reset/Start Request lines. Refer to the following table for input circuit specications.
Figure below illustrates how Remote Interlock inputs are factory-jumpered on the Quick Start Plug to
enable lasing for initial testing and troubleshooting purposes.
SYNRAD f201 Operator’s Manual Version 3.2
66
Page 69
technical reference
Pin 15Pin 9
User I/O connections
Figure below illustrates the input circuit’s equivalent internal schematic while following table provides
Firestar f201 input circuit specications.
Table 4-4 Input circuit specications.
Input Signal Name Input Device Type and Specications
PWM Input High-speed optoisolator LED, forward voltage drop (Vf) 1.5 VDC
O state Vmax +0.8 VDC
On state Vmin +3.5 VDC @ 3 mA
On state (continuous) Vmax +6.7 VDC @ 10 mA
Frequency, max. 100 kHz
Remote Reset/Start Request Bi-directional optoisolator LED, forward voltage drop (Vf) 1.15 VDC
Remote Interlock O state Vmax < 1.0 VDC
Shutter Open Request On state Vmin ±5.0 VDC @ 7 mA
On state (continuous) Vmax ±24.0 VDC @ 40 mA
Note: The Remote Reset/Start Request input must not be sent until Firestar’s +5 VDC power supply
has stabilized (approximately 200 ms after DC power-up).
Warning
The use of the Quick Start Plug bypasses the laser’s safety interlock
function, potentially exposing personnel in the area to invisible
serious
personal
injury
infrared laser radiation.
The Quick Start Plug is intended only for initial testing and troubleshooting by qualied personnel. In normal operation, the laser’s
Remote Interlock input should be connected to the machine’s safety
interlock circuitry.
Pin 1 PWM Return
Pin 9 PWM Input
Pin 3 Remote Interlock
Pin 4 +5 VDC
Pin 11 Input Common
Pin 12 Aux. DC Power Ground
Figure 4-8 Input equivalent schematic.
Rear (Solder Side)
of Male DB15
"Quick Start"
Shorting Plug
Pin 8Pin 1
PWM Input BNC
SYNRAD f201 Operator’s Manual Version 3.2
67
Page 70
technical reference
REMO
INPUT COMMON (11)
User I/O connections
Output signals
Firestar’s ve user outputs correspond to the status functions described below. Outputs are optoisolated,
bi-directional analog switches that allow for high-side or low-side switching. The shared connection,
Output Common, is separate from the laser’s chassis ground to allow high-side or low-side switching and
to isolate control signals for optimum EMI performance.
USER I/O INPUT SIGNAL PINS
PWM INPUT (+) (9)
PWM RETURN (–) (1)
220 Ohm, 1/8W
430 Ohm, 1/10W
TE RESET/START REQUEST (2)
REMOTE INTERLOCK (3)
SHUTTER OPEN REQUEST (10)
600 Ohm, 2W
600 Ohm, 2W
600 Ohm, 2W
FIRESTAR f-SERIES
INPUT CIRCUITRY
Figure 4-7 Quick Start Plug wiring diagram.
Firestar’s optically-isolated outputs are useful for sending laser status to a Programmable Logic Controller
(PLC) or computerized control system. Each of the ve outputs can source 50 mA at ±24 VDC maximum
for a total load of 250 mA. For controlling larger loads, use these outputs to drive a control relay.
Note: Laser Ready and Shutter Open outputs denote separate control functions. The Laser Ready
output (Ready LED On) may close while the Shutter Open output is open (Shutter LED O), but
the RF driver is disabled until both Laser Ready and Shutter Open outputs are closed (Ready
and Shutter indicators illuminated).
Pin 6 Laser Active
This bi-directional switched output is internally connected to Pin 13, Output Common, when
the laser is actively lasing (Lase indicator red). This output is open (high impedance) when no
beam is being emitted (Lase indicator O). Refer to the following table for output circuit specications.
Pin 7 Over Temperature
This bi-directional switched output is internally connected to Pin 13, Output Common, when
laser temperature is above safe operating limits (Temp indicator red). The output is open (high
impedance) when laser temperature is within operating limits (Temp LED green). After an over
temperature fault occurs, cool the laser and then cycle DC power to reset the laser. Refer to he
following table for output circuit specications.
SYNRAD f201 Operator’s Manual Version 3.2
68
Page 71
Pin 8 Laser Ready
This bi-directional switched output is internally connected to Pin 13, Output Common, when
the laser is enabled (Ready indicator On), indicating that lasing will occur when a PWM Command signal is applied to Pin 9 and Pin 1. When this output is initially switched closed, there is a
ve-second delay during which lasing is inhibited. This output is open (high impedance) when
the laser is disabled (Ready LED O ). Refer to Table below for output circuit specications.
Pin 13 Output Common
Use this pin to complete the return (ground) path for any output connection (Pin 6, 7, 8, 14, or
15). The Output Common line is protected by a 0.25 A self-resetting fuse.
Pin 14 Shutter Open
This bi-directional switched output is internally connected to Pin 13, Output Common, when
the Shutter Switch is Open and a Shutter Open Request signal is present (Shutter LED blue),
indicating that lasing may be enabled if other operating conditions are met. The output is open
(high impedance) when the Shutter Switch is Closed or the Shutter Open Request signal is
removed. When Shutter Open output closes (Shutter LED O), there is a ve-second delay until
PWM inputs are recognized. Refer to the below table for output circuit specications.
Pin 15 Interlock Open
This bi-directional switched output is internally connected to Pin 13, Output Common, when
remote interlock input circuitry is open (Remote Interlock LED red), indicating that lasing is
disabled. This output is open (high impedance) when the laser is enabled (Remote Interlock
indicator green). When this output is initially switched open, there is a ve-second delay during
which lasing is inhibited. See the following tables for output circuit specications.
technical reference
User I/O connections
Following gure illustrates the output circuit’s equivalent internal schematic and the following provides
Firestar
f201 output circuit specications.
Table 4-5 Output circuit specications.
Output Device Specications
Bi-directional MOSFET 2.5 Ohms Rdson
10 MOhms O
Voltage ±24 VDC, max.
Current 50 mA, max.
SYNRAD f201 Operator’s Manual Version 3.2
69
Page 72
technical reference
GROUND
enab
USER I/O PINS
TE INTERLOCK
Close switch to
USER I/O PINS
TE INTERLOCK
(11) INPUT COMMON
–12
User I/O connections
Sample I/O circuits
Sample inputs
Figure below illustrates one method of supplying a Remote Interlock signal using a customer-supplied
limit switch. Firestar’s +24 VDC Auxiliary Power output powers the circuit. Note that Pin 4, +5 VDC Auxiliary Power, could have been used instead, depending on circuit voltage requirements.
(5) +24 VDC AUXILIARY
Close switch to
le interlock
POWER
(3) REMO
(11) INPUT COMMON
(12) AUX. DC POWER
Figure 4-10 Customer-supplied interlock.
Figure below shows another variation for supplying a Remote Interlock signal to the laser. In this case,
the customer is using a limit switch and supplying a negative voltage to drive Firestar’s input circuit.
A Programmable Logic Controller (PLC) can also drive Firestar inputs. Figure below shows a typical
method for connecting to a PLC output module when only one Firestar input is used.
PLC
DC
OUTPUT
MODULE
+V
(+5–24V)
Figure 4-12 PLC driven interlock signal.
When multiple PLC outputs are used, connect Firestar inputs to the PLC as shown in Figure below. By
supplying voltage (+VDC) to Pin 11, Input Common, and pulling individual inputs to ground, each input
can be independently activated by the PLC’s output module.
USER I/O PINS
(3) REMOTE INTERLOC
(11) INPUT COMMON
PLC
DC
OUTPUT
MODULE
+V
(+5–24V)
Figure 4-13 Multiple PLC driven inputs.
USER I/O PINS
(11) INPUT COMMON
(2) REMOTE RESET/STA RT REQUEST
(3) REMOTE INTERLOCK
(10) SHUTTER OPEN REQUEST
SYNRAD f201 Operator’s Manual Version 3.2
71
Page 74
technical reference
(5)
PO
(6)
(12)
GR
(13) OUTPUT COMMON
USER I/O PINS
(5)
PO
(6)
(12)
GR
(13) OUTPUT COMMON
+V
USER I/O PINS
User I/O connections
Sample outputs
Firestar’s optoisolated, bi-directional switched outputs can drive small loads (50 mA max), PLC inputs, or
relays that can control higher current loads. Figure below illustrates one method of controlling a remote
warning lamp using power supplied by Firestar’s +24 VDC Auxiliary Power output. Remember to size
current-limiting resistor, R1, so that the current draw does not exceed 50 mA.
+24 VDC AUXILIARY
WER
LASER ACTIVE
AUX. DC POWER
OUND
R1
L
Figure 4-14 Firestar output driving warning lamp.
Figure below illustrates a method for controlling a higher voltage, higher current load by using a 24V
control relay. Ensure that the relay coil’s pull-in current does not exceed 50 mA. A diode or surge suppressor must be installed across the relay coil to prevent voltage spikes from damaging Firestar outputs.
+24 VDC AUXILIARY
WER
LASER ACTIVE
AUX.DC POWER
OUND
Figure 4-15 Firestar output driving relay.
SYNRAD f201 Operator’s Manual Version 3.2
72
L
Page 75
technical reference
(13)
(6)
(7)
(8)
(14)
(15)
User I/O connections
Figure below illustrates how Firestar’s outputs can drive the DC Input Module of a Programmable Logic
Controller (PLC). By supplying voltage (+VDC) to Pin 13, Output Common, each Firestar output is independently switched to activate individual PLC inputs.
Note: If you lengthen the DC Power cable, you must calculate and measure the additional voltage drop
to ensure that 96 VDC is available at the laser’s DC POWER connector under full-load conditions.
Depending on the additional length required, you may need to use larger gauge wire.
f201 DC Power cables
DC Power cables for f201 lasers are manufactured with #12 AWG wire; however, DC input current is split
between two positive (+) and two negative (–) conductors. Firestar f201 DC Power cables are available in
5 meter (16.4 ft) and 7 m (23 ft) lengths. The standard length is 5 meters. The minimum bend radius for
f201 DC Power cables must be greater than 124.5 mm (4.9 in).
F201 DC Power cables are wired into a Canon-type ME3476L14-4S plug that connects with a matching
receptacle on the laser’s rear panel. Table below provides pinout information for f201 DC Power cable
connectors.
Table 4-7 f201 DC Power connector pinouts.
Function Wire Color Code Terminal Designation
+96 VDC Red A
– VDC Black B
+96 VDC Red C
– VDC Black D
Ground Green Shell
SYNRAD f201 Operator’s Manual Version 3.2
74
Page 77
technical reference
Integrating Firestar safety features
The Integrating Firestar safety features section includes subsections:
■ Remote Interlock functions
Firestar’s DB-15 User I/O connector allows system integrators or end-users to integrate Firestar safety
features into their control system. Firestar’s keyswitch, shutter, and remote interlock functions serve
to enable or disable DC power to Firestar’s RF drive. Without DC power, the RF driver cannot supply RF
energy to the resonator, causing the CO2 gas to remain in a zero-energy state. Firestar status indicators
provide users with a quick visual indication of the laser’s operational status. All power to the laser’s RF
board is removed whenever Ready or Shutter indicators are O (Laser Ready or Shutter Open outputs
open).
OEM lasers
On OEM lasers, the Ready LED illuminates on DC power-up (provided that Shutter Open Request and
Remote Interlock inputs are enabled) and ve seconds later, DC power is applied to the RF driver. Over
temperature faults are reset by removing and then re-applying DC power after the laser has cooled. Remote interlock faults are not latched; the Ready LED illuminates yellow as soon as the interlock circuit is
closed (when the Remote Interlock LED turns from red to green) and ve seconds later lasing is enabled.
Although a Remote Reset/Start Request input is not needed to reset OEM faults, it can be used to inhibit
(disable) lasing. Disable the laser by applying a voltage in the range of ±5–24 VDC to Pin 2, the Remote
Reset/Start Request input. Removing voltage allows DC power to reach the RF driver and begins a
ve-second countdown after which lasing is enabled (Ready LED illuminates yellow). The RF driver is
disabled as long as voltage is applied to Pin 2.
Your control system can monitor the laser’s ready status on the User I/O connector by connecting your
system’s input between Pin 8, Laser Ready, and Pin 13, Output Common (see Figure 4-16).
SYNRAD f201 Operator’s Manual Version 3.2
75
Page 78
technical reference
Integrating Firestar safety features
The Laser Ready output closes when the laser is enabled (Ready LED illuminated yellow), indicating that
lasing is possible. The output is open (Ready LED o) when lasing is disabled.
Note: After the Laser Ready output closes, a ve-second delay occurs before lasing is enabled.
Remote interlock functions
Interlock circuits are often used to disable machinery when a shield, panel, or door is opened. Firestar’s
remote interlock function allows you to connect into an external remote interlock circuit and prevent
lasing by removing DC power from the laser’s RF driver boards when the circuit is electrically “open”.
Lasing is enabled when a Remote Interlock signal is present (Remote Interlock LED illuminated green),
if Ready and Shutter LEDs are illuminated, and disabled when the Remote Interlock signal is removed
(Remote Interlock LED red, Ready LED o). DC power is applied to the RF driver only when the Remote
Interlock LED is green and the Ready LED is yellow (and the Shutter LED is illuminated blue). Remote interlock functionality is provided by the Remote Interlock input signal on Pin 3 of the User I/O connector.
To use Firestar’s remote interlock feature, apply a voltage in the range of ±5–24 VDC to Pin 3, Remote Interlock. Applying an interlock signal causes the Remote Interlock LED to illuminate green and sends DC
power to the RF driver, which enables lasing after a ve-second delay (provided that the Ready LED is
yellow and the Shutter LED is blue). Removing the interlock signal removes DC power from the RF driver,
causing the Remote Interlock LED to turn red and the Ready LED to turn o. Lasing remains disabled
until a Remote Interlock signal is reapplied to Pin 3.
Your control system can monitor the laser’s remote interlock status on the User I/O connector by connecting your system’s input to Pin 15, Interlock Open, and Pin 13, Output Common (see Figure 3-16).
This output is closed when remote interlock circuitry is open (Remote Interlock LED illuminated red).
The output is open (in a high-impedance state) and the Remote Interlock LED is green when interlock
circuitry is closed.
■ Troubleshooting – explains how to troubleshoot common f201 problems.
SYNRAD f201 Operator’s Manual Version 3.2
81
Page 84
maintenance/
troubleshooting
Maintenance
The Maintenance section includes subsections:
■ Disabling f201 lasers
■ Daily inspections
■ Storage/shipping
■ Cleaning optical components
Disabling f201 lasers
Before performing any maintenance on your Firestar® f201 laser, be sure to completely disable the laser
by disconnecting the DC Power cable (or cables) from the rear of the laser.
Daily inspections
Perform the following steps daily to keep your Firestar f201 laser in optimum operating condition. Except for the procedures described below, no other service is required or should be attempted.
Caution
possible
equipment
If you operate your laser or marking head in a dirty or dusty environment, contact SYNRAD about the risks of doing so and precautions
you can take to increase the longevity of your laser, marking head,
and associated optical components.
damage
Warning
serious
personal
A risk of exposure to toxic elements, like zinc selenide, may result
when certain optical or beam delivery components are damaged. In
the event of damage to laser, marking head, or beam delivery optics,
contact SYNRAD, Inc. or the optics manufacturer for handling
instructions.
injury
1
Inspect all cooling tubing connections for signs of leakage. Check for signs of condensation that may
indicate the cooling water temperature has been set below the dew point temperature. Condensation
will damage electrical and optical components inside the laser. See Setting coolant temperature in the
Getting Started chapter for details on preventing condensation.
2
When using compressed air as a purge gas on your f201 laser, empty water traps and oil separa-
tors on each lter and/or dryer between the laser and your compressed air source. Compressed air
purity must meet the purge gas specications shown in Table 1-6 in the Getting Started chapter.
SYNRAD f201 Operator’s Manual Version 3.2
82
Page 85
maintenance/
troubleshooting
Maintenance
3
Inspect beam delivery components for signs of dust or debris and clean as required. When cleaning
the optical surfaces of beam delivery components, carefully follow the manufacturer’s instructions.
4
Visually inspect the exterior housing of the laser to ensure that all warning labels are present. Refer
to the Laser Safety chapter for f201 label types and locations.
Storage/shipping
When preparing the laser for storage or shipping, remember to drain cooling water from the laser. In cold
climates, any water left in the cooling system may freeze, which could damage internal components. After
draining thoroughly, use compressed shop air at no more than 29 PSI (wear safety glasses!) to remove any
residual water. When nished, cap all connectors to prevent debris from entering the cooling system.
When shipping SYNRAD lasers to another facility, we highly recommend that you ship the unit in its
original SYNRAD shipping container. If you no longer have the original shipping box and inserts, contact
SYNRAD Customer Service about purchasing replacement packaging. Refer to Packaging instructions in
the Technical Reference chapter for detailed instructions on properly packaging the laser for shipment.
Important Note: Failure to properly package the laser using SYNRAD-supplied shipping boxes and
foam/cardboard inserts as shown in the Packaging instructions may void the warranty. Customers may incur additional repair charges for shipping damage caused by
improper packaging.
Cleaning optical components
Debris or contaminants on the laser’s output coupler or external beam delivery components may aect laser processing and lead to damage or failure of the optics and/or the laser. Carefully follow the steps below
to inspect and clean the optical components in the beam path. Before beginning the cleaning process,
read this entire section thoroughly to ensure that all cleaning materials are available and that each step is
completely understood.
Caution
possible
equipment
damage
Because of their smaller beam diameter, Firestar lasers have signicantly higher power densities than previous SYNRAD lasers. This
means that any contamination on the laser’s output coupler (or on
any beam delivery optic) can absorb enough energy to damage one or
more optics in the beam path. Periodically inspect the laser’s output
coupler and all other beam delivery optics for signs of contaminants
and then carefully clean as required. In dirty environments, purge
laser optics using ltered air or nitrogen to prevent vapor and debris
from accumulating on optical surfaces.
Important – Firestar f201 and lasers do not have output windows
located inside the faceplate. NEVER disassemble the laser housing
to inspect f201 optics. Instead, always connect nitrogen or ltered
air to the laser’s Gas Purge port to prevent vapor and debris from
contaminating internal optical surfaces.
SYNRAD f201 Operator’s Manual Version 3.2
83
Page 86
maintenance/
troubleshooting
Maintenance
Danger
serious
personal
Ensure that DC power to the laser is turned o and locked out before
inspecting optical components in the beam path. Invisible CO2 laser
radiation is emitted through the aperture. Corneal damage or blindness may result from exposure to laser radiation.
injury
Important Note: Exercise great care when handling infrared optics; they are much more fragile than
common glass materials. Optical surfaces and coatings are easily damaged by rough
handling and improper cleaning methods.
Cleaning guidelines
■ Wear latex gloves or nger cots (powder-free) to prevent contamination of optical surfaces by dirt and
skin oils.
■ Never handle optics with tools; always use gloved hands or ngers.
■ Hold optics by the outer edge; never touch the coated surface.
■ Always place optics on lens tissue for protection; never place optics on hard or rough surfaces.
■ It may be necessary to use a cotton ball or ued cotton swab instead of a lens wipe to uniformly clean
the entire surface of small-diameter mounted optics.
■ Before using any cleaning agents, read Material Safety Data Sheets (MSDS) and observe all necessary
safety precautions.
Required cleaning materials
Table below lists the type and grade of materials required to properly clean optical surfaces.
Table 5-1 Required cleaning materials.
Cleaning Material Requirements
Latex gloves or nger cots Powder-free
Air bulb Clean air bulb
Ethyl or isopropyl alcohol Spectroscopic or reagent grade
Acetone Spectroscopic or reagent grade
Lens wipe (preferred) Optical (cleanroom) quality
Cotton balls or cotton swabs High-quality surgical cotton/high-quality paper-bodied
SYNRAD f201 Operator’s Manual Version 3.2
84
Page 87
maintenance/
troubleshooting
Maintenance
Cleaning optics
1
Shut o and lock out all power to the laser. You must verify that the laser is OFF (in a zero-energy
state) before continuing with the optical inspection!
2
Visually inspect all optical surfaces in the beam path, including the laser’s output coupler, for con-
taminants.
Caution
possible
lens
damage
3
Remove loose contaminants from the optic by holding a clean air bulb at an angle to the optic and
blow a stream of air at a glancing angle across the lens surface. Repeat as necessary.
4
Dampen a lens wipe with the selected cleaning agent. Alcohol (least aggressive) is best for initial
surface cleaning. Acetone (moderately aggressive) is best for oily residue or minor baked-on vapors
and debris.
Important Note: If acetone is used as a cleaning solvent, a second follow-up cleaning of the optical
surface using alcohol is required.
Do not allow the nozzle of the air bulb to touch the optical surface.
Any contact may damage the optic by scratching coatings on the
optical surface.
Do not use compressed shop air to blow contamination from the optic. Compressed air contains signicant amounts of water and oil that
form adsorbing lms on the optical surface.
Do not exert pressure on the surface of the optic during cleaning.
Optical surfaces and coatings are easily scratched by dislodged contaminants.
Use a new lens wipe on each pass as contaminants picked up by the
wipe may scratch the optical surface.
5
Gently, and without applying pressure, drag the damp lens wipe across the optical surface in a single
pass. Do not rub or apply any pressure, especially when using a cotton swab. Drag the wipe without
applying any downward pressure.
Note: Use a clean lens wipe on each pass. The wipe will pick up and carry surface contaminants that
may scratch optical surfaces or coatings.
To prevent streaking during the nal alcohol cleaning, drag the lens wipe slowly across the surface
so that the cleaning liquid evaporates right behind the wipe
6
Carefully examine the optic under a good light. Certain contaminants or damage such as pitting
cannot be removed. In these cases the optic must be replaced to prevent catastrophic failure.
7
Repeat Steps 4 through 6 as required, removing all traces of contaminants and deposits.
SYNRAD f201 Operator’s Manual Version 3.2
85
Page 88
maintenance/
troubleshooting
Troubleshooting
The Troubleshooting section includes subsections:
■ Introduction
■ Operational owchart
■ Functional block diagram
■ Status LEDs
■ Laser fault indications
■ Resetting faults
■ Other laser faults
■ Beam delivery optics
Introduction
This section is designed to help isolate problems to the module level only. Problems on circuit boards or
the laser tube are outside the scope of this guide because they are not user-serviceable assemblies; do not
attempt to repair them. Contact SYNRAD or a SYNRAD Authorized Distributor for repair or replacement information.
To troubleshoot Firestar f201 lasers, it is necessary to understand the sequence of events that must
happen before the laser can operate. Before attempting any service, we advise you to read the entire
troubleshooting guide and review both the operational owchart and the functional block diagram.
Danger
serious
personal
injury
This Class 4 laser product emits invisible infrared laser radiation in
the 10.6 µm CO2 wavelength band. Because exposure to direct or diffuse laser radiation can inict severe corneal injuries, always wear eye
protection when in the same area as an exposed laser beam. Do not
allow the laser beam to contact a person. This product emits an invisible laser beam that is capable of seriously burning human tissue.
Always be aware of the beam’s path and always use a beam block
while testing.
Caution
possible
equipment
Attempting repair of a SYNRAD Firestar laser without the express
authorization of SYNRAD, Inc. will void the product warranty.
If troubleshooting or service assistance is required, please contact
SYNRAD Customer Service.
damage
SYNRAD f201 Operator’s Manual Version 3.2
86
Page 89
maintenance/
troubleshooting
Troubleshooting
Operational owchart
The owchart in the gure below illustrates Firestar’s start-up sequence.
Laser Start Sequence
Apply 96 VDC
power to laser
Remote Interlock
indicator Green?
Ye s
Temp indicator
Green?
Ye s
Shutter indicator
Blue?
Ye s
Turn Keyswitch
to ON
(if equipped)
Ready indicator
Yellow?
No
No
No
No
Apply interlock signal
to Remote Interlock input
or install factory-supplied
Quick Start Plug
Check that cooling water
is flowing through laser
and that coolant temp
is below 22˚ C
Apply shutter open
signal to Shutter Open
Request input or install
factory-supplied Quick
Start Plug
Cycle Keyswitch from ON
to OFF/RESET and then
back to ON
(on OEM lasers,
cycle DC power)
Cycle DC power
(remove DC power,
wait 15 seconds,
reapply DC power)
Ensure that manual
Shutter Switch is set
to Open
Ready indicator
Yellow?
No
Figure 5-1 Operational owchart.
Ye s
Apply PWM Command
signal to laser
Lase indicator
illuminates Red to
indicate laser output
Ye s
SYNRAD f201 Operator’s Manual Version 3.2
87
Page 90
maintenance/
troubleshooting
Troubleshooting
Functional block diagram
Figure below is a functional block diagram illustrating the f201’s control architecture.
STATUS INDICATORS
LASE
SHUTTER
READY
TEMP
OVER
DETECTOR
TEMPERATURE
REMOTE
DETECTOR
INTERLOCK
REMOTE INTERLOCK
SHUTTER
CONTROL
LASE
INDICATOR
SHUTTER OPEN REQUEST
PWM INPUT
ISOLATION
INPUT
PWM OUT
CATGEN
DELAY TIMER
REMOTE RESET/START REQUEST
FAULT
TO RF DRIVER BOARD
TO CONTROL BOARD
REMOTE
INTERLOCK
LASER ACTIVE
INTERLOCK OPEN
SHUTTER OPEN
ISOLATION
OUTPUT
OVER TEMPERATURE
Figure 5-2 f201 functional block diagram.
SYNRAD f201 Operator’s Manual Version 3.2
88
LASER READY
CONNECTOR
DB15 USER I/O
POWER
SWITCH
INPUT
VOLTAGE
REGULATOR
DC POWER
READY
FAULT
DETECT
ON AND
96 VDC
Page 91
maintenance/
troubleshooting
Troubleshooting
Status LEDs
Firestar f201 LED indicators, also mirrored as output signals on the User I/O connector, provide status information to the user. Table below shows Firestar output signal and LED indicator states during normal
and fault conditions. User I/O outputs are Closed when the state indicated by the signal name is True.
Table 5-2 Status signals.
LED Status Output Signal Name User I/O Signal Status
LED Indicator Normal Fault Normal Fault
Remote Interlock Green – – Interlock Open Open – –
– – Red – – Closed
Temp Green – – Over Temperature Open – –
– – Red – – Closed
Shutter Blue – – Shutter Open Closed – –
– – O – – Open
Lase Red – – Laser Active Closed – –
– – O – – Open
When PWM Command pulses are applied (and are long enough to produce laser output) the Lase LED
illuminates red.
On OEM lasers, the Ready lamp illuminates on DC power-up when Remote Interlock and Temp indicators illuminate green. After the Ready indicator illuminates, a ve-second delay occurs before Firestar is
permitted to lase.
Note: Firestar Ready and Shutter indicators (as well as Laser Ready and Shutter Open outputs) denote
separate control functions. Although the Ready lamp may light while the Shutter LED is O (Shutter
Switch Closed or Shutter Open Request signal missing), no power is applied to the RF boards until both
Ready and Shutter indicators are illuminated.
SYNRAD f201 Operator’s Manual Version 3.2
89
Page 92
maintenance/
troubleshooting
Troubleshooting
Tables below show how 201 signal status changes when remote interlock, over-temperature, or shutter
closed conditions occur (fault conditions are shown in bold type).
Table 5-3 Normal operating condition.
LED Indicator LED Status Output Signal Output Status
Remote Interlock Green Interlock Open Open
Temp Green Over Temperature Open
Ready Yellow Laser Ready Closed
Lase (tickle active) O Laser Active Open
Lase (if PWM applied) Red Laser Active Closed
Table 5-4 Quick Start Plug or interlock/shutter inputs not connected.
LED Indicator LED Status Output Signal Output Status
Remote Interlock Red Interlock Open Closed
Temp Green Over Temperature Open
Ready O Laser Ready Open
Lase (tickle inactive) O Laser Active Open
Lase (if PWM applied) O Laser Active Open
Table 5-5 Interlock open condition.
LED Indicator LED Status Output Signal Output Status
Remote Interlock Red Interlock Open Closed
Temp Green Over Temperature Open
Ready O Laser Ready Open
Lase (tickle inactive) O Laser Active Open
Lase (if PWM applied) O Laser Active Open
SYNRAD f201 Operator’s Manual Version 3.2
90
Page 93
maintenance/
troubleshooting
Troubleshooting
Table 5-6 Over Temperature fault.
LED Indicator LED Status Output Signal Output Status
Remote Interlock Green Interlock Open Open
Temp Red Over Temperature Closed
Ready O Laser Ready Open
Lase (tickle inactive) O Laser Active Open
Lase (if PWM applied) O Laser Active Open
Table 5-7 Shutter closed condition.
LED Indicator LED Status Output Signal Output Status
Remote Interlock Green Interlock Open Open
Temp Green Over Temperature Open
Ready Yellow Laser Ready Closed
Lase (tickle inactive) O Laser Active Open
Lase (if PWM applied) O Laser Active Open
Table 5-8 VSWR fault.
LED Indicator LED Status Output Signal Output Status
Remote Interlock Green Interlock Open Open
Temp Green Over Temperature Open
Ready Yellow Laser Ready Closed
Shutter Blue (Flashing) Shutter Open Closed
Lase (tickle active) O Laser Active Open
Lase (if PWM applied)* Red Laser Active Closed
* A continuously ashing Shutter LED indicates a VSWR fault and the laser is limited to a maximum pulse length of 12 µs (ap-
proximately 6% duty cycle at 5 kHz) until the fault clears.
SYNRAD f201 Operator’s Manual Version 3.2
91
Page 94
maintenance/
troubleshooting
Troubleshooting
Laser fault indications
On f201 lasers manufactured after November 2007, Ready (RDY) and Shutter (SHT) LED’s ash to annunciate one of four laser faults. The following table lists error codes, the corresponding fault, and describes
basic corrective action. See the following section, Resetting faults, for detailed corrective actions.
Table 5-9 Laser error codes.
LED # of Blinks Fault Condition Corrective Action
RDY 1 blink Under Voltage Fault Verify 96 VDC (measured under load)/cycle power
RDY 2 blinks Over Voltage Fault Verify 96 VDC (measured under load)/cycle power
RDY 3 blinks RF Drive Switch Fault Remove DC power to the laser, wait 30 seconds, and
then reapply DC power
Resetting faults
Remote interlock condition
A remote interlock condition occurs when the Remote Interlock input opens (the Remote Interlock LED
changes from green to red).
To reset a remote interlock condition on Keyswitch-equipped lasers, re-establish the Remote Interlock
signal input (Remote Interlock LED changes from red to green) and cycle the Keyswitch from OFF/RESET
to ON (or apply a ±5–24 VDC Remote Reset/Start Request pulse with the Keyswitch set to ON). When
the Ready indicator illuminates, lasing is enabled after a ve-second delay.
Warning
serious
personal
injury
On Firestar f201 OEM lasers, remote interlock faults are not latched.
Clearing the fault condition re-enables the Ready indicator and the
laser will re after the ve-second delay provided that the Shutter indicator is lit and a PWM Command signal is applied. Because
exposure to 10.6 µm CO2 laser radiation can inict severe corneal injuries and seriously burn human tissue, the OEM or System Integrator must ensure that appropriate safeguards are in place to prevent
unintended lasing.
SYNRAD f201 Operator’s Manual Version 3.2
92
Page 95
maintenance/
troubleshooting
Troubleshooting
On OEM lasers, remote interlock conditions are not latched. Re-establish the Remote Interlock signal
input (Remote Interlock LED changes from red to green) to enable the Ready indicator and begin lasing
after the ve-second delay.
Over temperature fault
Over temperature faults occur when thermal limits in the laser are exceeded (the Temp LED changes from
green to red).
To reset an over-temperature fault on Keyswitch-equipped lasers, lower coolant temperature below
°C (f201), cycle DC power to the laser, and then cycle the Keyswitch from OFF/RESET to ON (or apply a
±5–24 VDC Remote Reset/Start Request pulse with the Keyswitch set to ON). When the Ready indicator
illuminates, lasing is enabled after a ve-second delay.
Note: Because of the over-temperature latch circuit, the Temp indicator remains red prior to cycling
power even after the laser has cooled suciently to begin operation. If the Temp indicator remains
red after cycling power, the laser is not suciently cooled—cool the laser for several more minutes, cycle power, and then cycle the Keyswitch.
Over-temperature faults are now latched. To reset an over-temperature fault, lower coolant temperature
below 28 °C (f201) to cool the laser and then cycle DC power to the laser. Once the Temp indicator is
green and the Ready lamp is lit, lasing is possible after the ve-second delay.
Under voltage fault
An under voltage fault occurs when the DC input voltage falls below preset limits. This fault is indicated
by the Ready LED ashing 1 blink. To reset an under voltage fault, rst correct the voltage problem and
ensure that 96 VDC is measured at the laser under full-load conditions. Next, cycle DC power o and then
on again. On Keyswitch-equipped lasers, you must also toggle the Keyswitch from OFF/RESET to ON (or
apply a ±5–24 VDC Remote Reset/Start Request pulse with the Keyswitch set to ON). When the Ready
LED illuminates, lasing is enabled after a ve-second delay.
Over voltage fault
An over voltage fault occurs when the DC input voltage rises above preset limits. This fault is indicated by
the Ready LED ashing 2 blinks. To reset an over voltage fault, rst correct the voltage problem and ensure
that 96 VDC is measured at the laser under full-load conditions. Next, cycle DC power o and then on
again. On Keyswitch-equipped lasers, you must also toggle the Keyswitch from OFF/RESET to ON (or apply
a ±5–24 VDC Remote Reset/Start Request pulse with the Keyswitch set to ON). When the Ready LED
illuminates, lasing is enabled after a ve-second delay.
SYNRAD f201 Operator’s Manual Version 3.2
93
Page 96
maintenance/
troubleshooting
Troubleshooting
RF Drive Switch fault
An RF Drive DC fault occurs on power-up when the tube fails to breakdown or a fault occurs in the RF
Driver’s 96-volt switching circuitry. In this case, the Ready LED ashes 3 blinks. If an RF Drive DC fault
occurs, reset the laser by removing DC power to the laser, wait 30 seconds, and then reapply power. On
Keyswitch-equipped lasers, you must also toggle the Keyswitch from OFF/RESET to ON (or apply a ±5–24
VDC Remote Reset/Start Request pulse with the Keyswitch set to ON) When the Ready LED illuminates, lasing is enabled after a ve-second delay. If the RF Drive DC fault reappears, contact SYNRAD or
a SYNRAD Authorized Distributor.
PWM Drive/RF Oscillation faults
New circuitry on the f201 RF board has eliminated PWM Drive and RF Oscillation faults. The 4-blink fault
indication no longer exists.
VSWR fault
A VSWR fault indicates an impedance mismatch between the tube and RF Driver and occurs when a
PWM Command signal is rst applied. This fault is annunciated by the Shutter LED ashing continuously
until the fault disappears. If a VSWR fault occurs, the VSWR circuit limits the maximum PWM pulse
length to 12 microseconds (µs). This equates to a 6% maximum duty cycle at a 5 kHz frequency or a 24%
maximum duty cycle at a 20 kHz PWM frequency. When the VSWR fault clears, the laser resumes normal
operation without cycling DC power. If the VSWR fault fails to clear or reappears, contact SYNRAD or a
SYNRAD Authorized Distributor.
The leading causes of VSWR failures are environmental conditions, like cold overnight temperatures
while the laser is o. In cases like this, it may take as long as 60 seconds for tickle to breakdown the tube
and begin normal daily operation.
Other laser faults
Each Symptom listed below describes a particular fault condition. For each Symptom, specic causes and
solutions are described in the Possible Causes section.
Symptom:
■A remote interlock condition is indicated by the following status LEDs and I/O states:
Remote Interlock LED – Red Interlock Open output – Closed
Temp LED – Green Over Temperature output – Open
Ready LED – O Laser Ready output – Open
Lase LED – O Laser Active output – Open
SYNRAD f201 Operator’s Manual Version 3.2
94
Page 97
maintenance/
troubleshooting
Troubleshooting
Possible Causes:
■No voltage is applied to Pin 3 (Remote Interlock) of the User I/O connector.
Warning
serious
personal
injury
On systems using remote interlocks, check to see that a positive or negative voltage in the range of ±5–24
VDC is applied to Pin 3, Remote Interlock, with respect to Pin 11, Input Common, on the User I/O connector (refer to User I/O connections in the Technical Reference chapter for details). For systems not using
interlocks, connect the factory-supplied Quick Start Plug to the User I/O connector on the laser’s rear panel
or wire your male DB-15 connector so that Pin 11 (Input Common) is jumpered to Pin 12 (Auxiliary DC
Power Ground) and Pin 3 (Remote Interlock) is jumpered to Pin 4 (+5 VDC Auxiliary Power).
On Firestar f201 OEM lasers, remote interlock faults are not latched.
Clearing the fault condition re-enables the Ready indicator and the
laser will re after the ve-second delay provided that the Shutter indicator is lit and a PWM Command signal is applied. Because
exposure to 10.6 µm CO2 laser radiation can inict severe corneal injuries and seriously burn human tissue, the OEM or System Integrator must ensure that appropriate safeguards are in place to prevent
unintended lasing.
Symptom:
■An over-temperature fault is indicated by the following status LEDs and I/O states:
Remote Interlock LED – Green Interlock Open output – Open
Temp LED – Red Over Temperature output – Closed
Ready LED – O Laser Ready output – Open
Lase LED – O Laser Active output – Open
Possible Causes:
■ Coolant temperature is above 28 °C (f201) or there is inadequate coolant ow through the laser.
Check that your chiller is maintaining a water temperature between 18 °C–28 °C (64 °F–82 °F) at a ow
rate of 9.5 lpm (2.5 GPM) for f201.
If water temperature is OK, check the ow rate. The simplest way to do this, if a ow meter is not available, is to disconnect the cooling tubing from the chiller inlet (or the drain) and run the cooling water for
one minute into a ve-gallon bucket; you should have close to the number of gallons listed above. If there
is much less than the recommend ow, check the cooling path for kinked or pinched cooling tubes or
check the chiller for a clogged or dirty lter.
SYNRAD f201 Operator’s Manual Version 3.2
95
Page 98
maintenance/
troubleshooting
Troubleshooting
On f201 lasers, the over-temperature fault (indicated by the Temp indicator turning red) is latched. This
means that if an over-temperature condition occurs the Temp indicator will turn red, the Over Temperature output will Close, the Ready light goes out, and lasing is disabled. Because of its latched condition,
the Temp indicator will remain red even after the laser has cooled suciently to begin operation.
To reset an over-temperature fault, lower coolant temperature below 28 °C (f201) and then cycle DC
power (remove DC power, wait 30 seconds, reapply DC power). On Keyswitch-equipped f100/f200 lasers
you must also cycle the Keyswitch from OFF/RESET to ON (or with the Keyswitch ON, apply a positive or
negative voltage in the range of ±5–24 VDC to Pin 2, Remote Reset/Start Request) after re-applying DC
power. When the Ready indicator illuminates, lasing is enabled after the ve-second delay. If the Temp
indicator remains red after cycling power, continue to ow cooling water through the laser for a few
more minutes and then cycle DC power again.
Symptom:
■ Your OEM laser has quit lasing or lasing halted and then restarted. The Lase LED may be O or On
depending on whether PWM Command signals are being applied, but no fault is indicated.
Remote Interlock LED – Green Interlock Open output – Open
Temp LED – Green Over Temperature output – Open
Ready LED – Yellow Laser Ready output – Closed
Lase LED – O or On Laser Active output – Open or Closed
■ The remote interlock circuit momentarily opened.
Remote interlock faults are not latched on OEM lasers. This means that if an interlock open fault occurs,
the Remote Interlock indicator will turn red, the Interlock Open output will Close, the Ready light goes out,
and lasing is disabled. However, if the interlock circuit closes, the Remote Interlock indicator will change
from red to green, the Ready light illuminates, and ve seconds later lasing is enabled.
Symptom:
■ There is no output laser beam and all LED status indicators are O
Possible Causes:
■ The 20 A fast-acting fuse has blown.
Remove DC power from the laser and then check each fuse on the rear panel. Replace as required using
only a Bussman ABC-20 or equivalent.
Symptom:
■ There is no output laser beam from your f201 laser, but status indicators appear as:
Remote Interlock LED – Green Interlock Open output – Open
Temp LED – Green Over Temperature output – Open
Ready LED – O Laser Ready output – Open
Lase LED – O Laser Active output – Open
SYNRAD f201 Operator’s Manual Version 3.2
96
Page 99
maintenance/
troubleshooting
Troubleshooting
Possible Causes:
■ One of the two 20 A fast-acting fuses has blown.
On f201 lasers, a single blown 20 A fuse on the laser’s rear panel may cause the Ready lamp to remain
O while other LEDs indicate normal status. Remove DC power from the laser and then check each fuse
on the rear panel. Replace as required using only a Bussman ABC-20 or equivalent.
Beam delivery optics
Troubleshooting
Symptom:
■ The laser loses power over time; laser output power must be increased to maintain performance.
Possible Causes:
■ Beam delivery optics are coated by vapor residue or debris.
Shut down the laser and carefully inspect each optic in the beam delivery path, including the laser’s output coupler. Remember that optics are fragile and must be handled carefully; preferably by the mounting ring only. If the optic requires cleaning, refer back to Maintenance for cleaning instructions. Use only
recommended cleaning materials (see Table 5-1) to prevent scratching delicate optical surfaces.
If the focusing optic is pitted, it must be replaced immediately. Because of the extremely high power density of Firestar lasers, pits or debris on the lens surface may absorb enough energy from the focused beam to
crack the lens. If this happens, other optics in the beam path may be contaminated or damaged as well.
Warning
serious
personal
injury
The use of aerosol dusters containing dioroethane causes “blooming”, a condition that signicantly expands and scatters the laser
beam. This beam expansion can eect mode quality and/or cause
laser energy to extend beyond the connes of optical elements in
the system, possibly damaging acrylic safety shielding. Do not use air
dusters containing dioroethane in any area adjacent to CO2 laser
systems because dioroethane persists for long time periods over wide
areas.
SYNRAD f201 Operator’s Manual Version 3.2
97
Page 100
maintenance/
troubleshooting
Troubleshooting
Caution
possible
equipment
If you operate your laser or marking head in a dirty or dusty environment, contact SYNRAD about the risks of doing so and precautions
you can take to increase the longevity of your laser, marking head,
and associated optical components.
damage
When operating f201 lasers, connect nitrogen or breathing-grade air to the Gas Purge connector.
Purging the laser creates a positive pressure inside the laser housing that prevents dirt and debris from
accumulating on optical surfaces including beam combiner, telescope, and turning optics. In condensing atmospheres, a gas purge helps to reduce the potential for condensation damage. See Connecting
in the Getting Started chapter for information on using a gas purge. Table 2-6 in the Getting Started
chapter provides ltering and drying specications for recommended purge gases.
Danger
serious
personal
Ensure that DC power to the laser is turned o and locked out before
inspecting optical components in the beam path. Invisible CO2 laser
radiation is emitted through the aperture. Corneal damage or blindness may result from exposure to laser radiation.
injury
In addition to a gas purge for the laser housing, many laser processing applications like cutting, welding, and drilling require an assist gas for the focusing/gas jet assembly. When the application requires
air (instead of nitrogen) as an assist gas, we recommend the use of breathing quality air available in
cylinders from a welding supply company. Because compressed shop air contains minute particles of oil
and other contaminants that will damage optical surfaces, it must be carefully ltered and dried before
use as a purge or assist gas. Refer to Table 4-1, Assist gas purity specications, in the Technical Reference
chapter for a list of common assist gases and their corresponding ltering and drying specications.
Warning
serious
personal
A risk of exposure to toxic elements, like zinc selenide, may result
when certain optical or beam delivery components are damaged. In
the event of damage to laser, marking head, or beam delivery optics,
contact SYNRAD, Inc. or the optics manufacturer for handling
instructions.
injury
SYNRAD f201 Operator’s Manual Version 3.2
98
Loading...
+ hidden pages
You need points to download manuals.
1 point = 1 manual.
You can buy points or you can get point for every manual you upload.