Simpson Electric F45 Instruction Manual

Falcon F45 Series Digital Panel Meter
2
m
Full 4-1/2 Digit, Bright Red 0.56" (14.2mm) Display
Broad Range Display Scaling
Short 2.94" (74.7mm) Deep, 1/8 DIN Case
Screw Terminal Connector for Easy Installation
Four User-Settable Ranges: 200mV, 2V, 20V, 200V
One Factory-Settable Range: 750V
Optional Isolated 9-32VDC Power Supply
DC Voltage
The Falcon Series digital indicators are premium quality 1/8 DIN meters for industrial applications. All Falcon units feature jumper-selectable decimal point (internal and on the connector for remote decimal point) and display scal­ing, providing wide application flexibility. In addition, signal input ranges are easy to change with jumpers on the main board. The Falcon has a 0.56" bright red LED display for high visibility.
4.04"
102.5mm
3.59"
91.2mm
1.77"
45mm
3.62"
91.9mm
Compactly designed for applications requiring minimal rear panel depth, the Falcon fits a standard 1/8 DIN panel cut­out (91.9mm x 45mm) and requires less than 3" behind the panel. A screw terminal connector is a standard feature for easy wiring of the power supply and signal input connections.
Engineering Label
3.78"
96mm
1.89"
48mm
0.44"
11.2mm
2.94"
74.7mm
1.74"
44.2mm
2.2
56.5
Mounting Requirements
The Falcon series 1/8 DIN indicators require a panel cutout of 1.77" (45mm) high by 3.62" (91.9mm) wide. To install the Falcon into a panel cutout, remove the clips from the side of the meter. Slide the meter through your panel cutout, then slide the mounting clips back on the meter. Press evenly to ensure a proper fit.
Engineering Label Placement
If replacement of the engineering unit label is required, place the tip of a ball-point pen into the small hole at the base of the engineering label in the bezel. Slide the label up until it pops out. Grasp and remove. Slide the new label half the distance in, then use the ball-point pen to slide it down into place.
Specifications
DISPLAY Type: 7-segment, red LED Height: 0.56" (14.2mm) Decimal Point: 4-position programmable
internally or at terminal block J112 Overrange indication: most significant digit = “1”; other digits blank Polarity: Automatic, with "-" indication, "+" indication implied
POWER REQUIREMENTS AC Voltages: 120 or 220VAC, ±10%
50/60Hz
DC Voltages: 9-32VDC; 9V -1% and 32+1% Power Consumption: 2VA
ACCURACY @258C
±0.02% of reading ±1 count 750 ±2 counts
ENVIRONMENTAL Operating Temperature: 0 to 55ºC Storage Temperature: -10 to 60ºC Relative Humidity: 0 to 85% non-condensing
Temperature Coefficient:
(±0.05% of input ±0.1 digit)/C
Warm-up Time: Less than 15 minutes Response Time: Less than 1 second
NOISE REJECTION NMRR: 60dB, 50/60Hz CMRR: (w/1k unbalanced @60Hz):
90dB min.
ANALOG TO DIGITAL CONVERSION Technique: Dual slope integration Rate: 2.5 samples per second, nominal
MECHANICAL Bezel: 3.78" x 1.89" x .44"
(96 x 48 x 11.2mm)
Depth: 2.94"(74.7mm) Panel Cut-out: 3.62" X 1.77"
(91.9 x 45mm 1/8 DIN)
Case Material: 94V-1, UL rated Noryl® Weight: 9.0oz (255.1g)
INPUTS: DC Voltage
Input Display Input Maximum Range Resolution Impedance Overload
200mV 10V 100M 50V 2V 100V 10M 100V 20V 1mV 10M 100V 200V 10mV 10M 250V
Wiring Diagram
2
1
IN HI
IN LO
Input Signal: Connect the signal to be monitored to the IN HI and IN LO terminals. These are terminals #1 and #2.
Supply Power: Connect the power to terminals #11 and #12. Note that if AC power is applied, terminal #11 is for neutral, and terminal #12 is for hot. If DC power is used, terminal #11 is for -DC, and #12 is for +DC.
Display Hold: This feature allows you to hold the displayed value indefinitely. A remote switch or computer, etc. can be used to activate this feature. To activate feature, short circuit terminal block J112, pins 3 and 4 (Hold and DIG GND). This connection must be kept isolated from other circuitry. To hold multiple units, separate poles of the switch must be used to maintain the isolation.
These instruments are designed for maximum safety to the operator when
mounted in a panel according to instructions. They are not to be used
unmounted or for exploratory measurements in unknown circuits.
3
HOLD
45
+REF
-REF
DP 1
67
DP2
DP3
8
DP C
10 11 12
9
+EXC
-EXC
AC GND
AC LO
-DC
AC HI
+DC
Decimal Point Selection
From terminal block J112: The decimal point can be set from the rear screw terminal block J112. Connect the appropriate DP point (DP 1, 2, 3, 4) to the DIG GND terminal. Internal jumper (J107) should be removed and stored on the last contact of J107.
Point Connect
Decimal
1.9999 DIG GND to DP1
19.999 DIG GND to DP2
199.99 DIG GND to DP3
750.0 DIG GND to DP4
From main board: The decimal point can also be selected by accessing the main board. Move the push-on jumper J107 across the correct letter.
2345
1
IN HI
IN LO
HOLD
6789
DP2
DP3
DP 1
+REF
-REF
Decimal Jumper
Point Position at J107
1.9999 D
19.999 C
199.99 B
750.0V A
DP C
10 11 12
+EXC
-EXC
AC GND
AC HI
+DC
AC LO
-DC
Voltage Range Selection
All Falcon Indicators are configured initially per the customer specified part number. Range changes can easily be accomplished as follows: Disconnect power and pop the front bezel off with a small screwdriver, taking care to keep the gaskets in place. Unscrew the main board from the case with a Phillips head screwdriver, and slide the main board out. Note: If a new range is selected, the calibration procedure must also be per­formed. Only perform this section if a different function or range is required.
Input J105 J106 J101 J102 J103 JU101* Range PJ PJ PJ PJ PJ Jumper Position 200mV C NO NO YES NO V 2V A NO YES NO NO V 20V B YES YES NO NO V 200V D NO YES NO NO V
* 750 Volt range may be configured upon order by factory or Authorized Service Center
Note: JU101 is a hard wire jumper and can be removed by cutting it.
Resoldering the JU jumpers is not recommended. If this is required, or if a function is to be changed from volts to current, Simpson recommends returning the Falcon to the factory or an Authorized Service Center. After moving the jumpers to the desired location, put the Falcon back together and install in your panel, or proceed to calibration.
Note: See diagram in Decimal Point Selection panel for J106 jumper location.
Example: 200mV input
Display Scaling
The Falcon display can be easily scaled to a broad range of engineering units. The meter can be scaled down to the front bezel with a small screwdriver. The scaling procedure is per­formed at terminal J1 located on the left side of the display board. There are four sets of pins of J1. Each group of pins programs a particular parameter of the scaling procedure when a push-on jumper (supplied with the Falcon) is momentarily placed across a set of pins.
Before scaling the parameters, begin by disconnecting power to the Falcon. Place push-on jumper onto “SET UP” pins C and 2, then apply power. At this point, the Falcon performs a self test (scrolls LED display) and then indicates LoE, allowing you to enter the low electrical input value.
The scaling parameters for electrical low input value (LoE), electrical high input value (HiE), display low value (Lo), and display high value (Hi) are performed as follows:
Remove push-on jumper from C-2 and then replace. Numeral zero or digit of LoE value (set at the factory) and flashing decimal point are displayed at starting LED position. The digit to be entered is always to the left of the flashing decimal point. The flashing decimal point is moved one position each time the push-on jumper is momentarily placed across “set up” pins C and 2. Change the value at the LED position chosen by momentarily placing jumper across “up” pins C and 3. This will increase the value by 1 for each repetition. Repeat jumper placement until the desired value is displayed. To decrease a displayed value, momentarily place jumper across “DOWN” pins C and 4.
After the desired numeric value is obtained, remove the jumper and momentarily place it across the “ENTER” pins C and 1. This will store the value in memory and move on to the next parameter. Repeat this procedure through each parameter.
After programming the last parameter (Hi) of the scaling procedure, LED will display “END” and then indicate scaled value of the electrical input to the meter. The unit is now in run mode.
Note: During programming, the flashing decimal point is considered as a cursor only. The placement of the digits in the positions 1 to 5 is important, and dependent on the set input range.
1
/2of the input value. Remove
Enter-Saves digits displayed and advances to next scaling parameter when jumper makes momentary contact with pins 1 and C.
UP-Changes digit displayed to next ascend­ing value when jumper makes momentary contact with pins 3 and C.
DOWN-Changes digit displayed to next descending value when jumper makes momentary contact with pins 4 and C.
SET UP-Starts scaling cycle of the four scaling parameters (LoE, HiE, Lo and Hi) and selects digit to be entered to the left of the flashing decimal point on display when jumper makes momentary contact with pins 2 and C.
Typical Scaling Application
50mV Shunt; NOTE: 200mV range is used
Electrical Characteristics Falcon LED Display
Current: 100 amps LoE: 0 Lo: 0 Voltage Drop: 50mV HiE: 50.00 Hi: 250.0
Excitation Output
An optional feature of the Falcon is a 12 or 24VDC output. The Falcon is set at the factory per your order to include an excitation plug-in printed circuit board. This PCB, located on the main board, can be reconfig­ured by moving push-on jumper J201 to change the excitation output.
J201 Push-On Jumper
(Component Side)
24V Set-up
• 24V • • 12V •
12V Set-up
• 24V • • 12V •
Solder
Side
Excitation
PCB
J108
Calibration
The unit is calibrated at the factory per order. If you selected another range and moved the jumpers, your Falcon will need to be recalibrated. If parameters (HiE, etc.) are changed, scaling must be performed prior to calibration.
1) Remove the bezel with a small screwdriver.
2) Apply a full scale input signal to the instrument.
3) Adjust the potentiometer R1 located on the upper left-hand side of the display board until display shows the full scale voltage.
4) Replace bezel carefully.
Application Example
A manufacturing plant requires a low cost digital meter to replace an analog panel meter on a 150VDC motor. The upgrade is required because the operator requires a display hold feature that is not available on the analog model.
A Falcon 3-1/2 digit indicator (200DCV) can fit this application. The meter needs no additional scaling before being installed into the panel. A switch is required for the display hold option, as one is not supplied by Simpson.
The meter is installed in parallel with the motor like the analog meter, and is ready to be placed in the panel. The Falcon is less than 3" deep, fitting well into the restricted panel space. It will take up about as much space as the analog meter it replaced. In addition, display hold is a standard feature on the Falcon. By shorting connections #3 and #4, the operator can hold the display to take a reading, and then remove the short to reactivate the indication mode. A switch can be used to short the connections.
The meter is shipped ready to install, keeping down-time to a minimum.
Motor
Falcon 200VDC
Supply
Ordering Information
F45 Basic Unit
F45
4-1/2 Digit Indicator
1 2 3 4
*Meets CE EMI EN-50082-1, EN-55022, EN-61000-3-2, EN-61000-3-3
Power Supply
120VAC 220VAC* 9-32VDC 120VAC*
Accessories
Portable
Switchboard
Range
200mV
11 12 13 14
External shunts enable digital panel meters to indicate higher DC currents than can be provided with self contained internal shunt meters. The shunt is installed in series with the load and the source. The shunts provide a DCmV drop which is sent to the display unit. The Falcon can be scaled to display the actual current load between the load and the source. Simpson offers por­table and switchboard shunts.
Each portable shunt includes 5’ leads.
2V
20V
200V
Excitation
0
1 2
None 12VDC 24VDC
Safety Symbols
Order Information
Portable Shunts
( 50mV )
Amps Cat. Number
1 06700 5 06703 10 06704 15 06705 25 06707 30 06708 50 06709 75 06711 100 06713 150 06714 200 06715
The WARNING sign denotes a hazard. It calls attention to a procedure, practice, or the like, which, if not correctly performed or adhered to, could result in personal injury.
The CAUTION sign denotes a hazard. It calls attention to an operating procedure, practice, or the like, which, if not cor­rectly adhered to, could result in damage to or destruction of part or all the instrument.
Switchboard Shunts
( 50mV )
Amps Cat. Number
100 06500 150 06503 200 06504 250 06505 300 06506 400 06507 500 06508
SIMPSON ELECTRIC COMPANY 520 Simpson Avenue, Lac du Flambeau, WI 54538 • (715) 588-3311 • FAX (715) 588-3326
www.simpsonelectric.com Printed in U.S.A. Part No. 6-115912, Edition 16, 06/08