Simpson Electric F35 Instruction Manual

2
m
Full 3-1/2 Digit, Bright Red 0.56" (14.2mm) Display
Broad Range Display Scaling
Short 2.94" (74.7mm) Deep, 1/8 DIN Case
Screw Terminal Connector for Easy Installation
Three User-Selectable Ranges: 4-20mA, 1-5V, 0-10V
Jumper-Selectable Decimal Point
Optional Isolated 9-32VDC Power Supply
Optional Excitation Output of 12VDC or 24VDC
DC Process
The Falcon Series digital indicators are premium quality 1/8 DIN meters for industrial applications. All Falcon units feature jumper-selectable decimal point (internal and on the connector for remote decimal point) and display scal­ing, providing wide application flexibility. In addition, signal input ranges are easy to change with jumpers on the main board. The Falcon has a 0.56" bright red LED display for high visibility.
Installation and Panel Cutout
4.04"
102.5mm
3.59"
91.2mm
1.77"
3.78"
96mm
1.89"
48mm
0.44"
11.2mm
45mm
91.9mm
2.94"
74.7mm
3.62"
1.74"
44.2mm
2.2
56.5
Compactly designed for applications requiring minimal rear panel depth, the Falcon fits a standard 1/8 DIN panel cut­out (92mm x 45mm) and requires less than 3" behind the panel. A screw terminal connector is a standard feature for easy wiring of the power supply and signal input con­nections.
Engineering Label
Mounting Requirements
The Falcon series 1/8 DIN indicators require a panel cutout of 1.77" (45mm) high by 3.62" (92mm) wide. To install the Falcon into a panel cut­out, remove the clips from the side of the meter. Slide the meter through your panel cutout, then slide the mounting clips back on the meter. Press evenly to ensure a proper fit.
Engineering Label Placement
If replacement of the engineering unit label is required, place the tip of a ball-point pen into the small hole at the base of the engineering label in the bezel. Slide the label up until it pops out. Grasp and remove. Slide the new label half the distance in, then use the ball-point pen to slide it down into place.
Specifications
DISPLAY Type: 7-segment, red LED Height: 0.56" (14.2mm) Decimal Point: 3-position programmable,
internally or on the terminal block Overrange indication: most significant digit = “1”; other digits blank
Polarity: Automatic, with "-" indication,
ENVIRONMENTAL Operating Temperature: 0 to 55ºC Storage Temperature: -10 to 60ºC Relative Humidity: 0 to 85% non-condensing Temperature Coefficient:
(±0.01% of input ± 0.05 count)/ºC
Warm-up Time: Less than 15 minutes Response Time: Less than 1 second
"+" indication implied
NOISE REJECTION POWER REQUIREMENTS AC Voltages: 120 or 220VAC, ±10% 50/60Hz DC Voltages: 9-32VDC, ±1%
NMRR: 50dB, 50/60Hz
CMRR: (w/1K unbalanced @ 60Hz):
90dB min.
Power Consumption: 3VA
ANALOG TO DIGITAL CONVERSION ACCURACY @25º C
±0.1% of reading ±1 count
Technique: Dual slope integration
Rate: 3 samples per second, nominal
Wiring Diagram
2
1
IN HI
IN LO
Input Signal: Connect the signal to be monitored to the IN HI and IN LO terminals. IN HI is terminal #1, IN LO is terminal #2.
Supply Power: Connect the supply power to terminals #11 and #12. Note that if AC power is supplied, terminal #11 is for Neutral, and terminal #12 is for Hot. If DC power is used, terminal #11 is for
-DC, and #12 is for +DC.
Display Hold: This feature allows you to hold the displayed value indefinitely. A remote switch can be used to make the connection. To activate the display hold, short terminal blocks #3 and #4 (Hold Ref and +Ref). This connection must be kept isolated from other circuitry. To hold multiple units, separate poles of the switch must be used to maintain the isolation.
45
3
+REF
HOLD
-REF
These instruments are designed for maximum safety to the operator when mounted in a panel according to instructions. They are not to be used unmounted or for exploratory mea­surements in unknown circuits.
DP 1
67
DP2
DP3
8
DP C
10 11 12
9
+EXC
-EXC
AC GND
AC LO
-DC
AC HI
+DC
MECHANICAL Bezel: 3.78"x 1.89" x .44"
(96 x 48 x 11.2mm)
Depth: 2.94" (74.7mm) Panel Cut out: 3.62" X 1.77"
(91.9 x 45mm 1/8 DIN)
Case Material: 94V-1, UL rated Noryl® Weight: 9.0oz (255.1g)
INPUTS: DC Process
Input Display Voltage Maximum Range Resolution Drop Overload
4-20mA 10A 0.2V 100mA (Input Impedance) 1-5V 10mV 10M 100V 0-10V 10mV 10M 100V
Decimal Point Selection
From terminal block: The decimal point can be set from the rear
screw terminal block by connecting the appropriate decimal point (DP 1, 2, 3, ) to the DP C terminal. The J105 jumper must be in the D position (see diagram under "From front panel").
Decimal Point Connect
1.999 DP C to DP1
19.99 DP C to DP2
199.9 DP C to DP3 1999 No Decimal
From front panel: The decimal point can also be selected by removing the front bezel from the meter. Move the push-on jumper J105 across the correct letter.
Decimal Jumper Point Position at J105
1.999 A
19.99 B
199.9 C 1999 D* * No Decimal Point
2
45
3
1
IN HI
IN LO
HOLD
-REF
+REF
6789
DP2
DP 1
DP3
DP C
10 11 12
+EXC
-EXC
AC GND
AC LO
-DC
AC HI
+DC
Before switching the instrument on, make sure the supply volt­age matches the power source required of the instrument as indicated on the hook-up label affixed to the instrument.
Process Range Selection
All Falcon Indicators are configured initially per the customer specifica­tions. Range changes can be accomplished as follows:
Disconnect power from the unit. Remove the unit from the panel. Remove the front bezel by inserting slotted screwdriver in the vertical slots on either side of the bezel and then turning to pry the bezel off. Unscrew the two Phillips head screws at either side of the circuit board. Finally, push on the green connector assembly in the back of the unit to slide the main circuit board out from the meter. Change jumpers according to the chart below.
Note: If a new range is selected, the calibration procedure must also be performed.
Range J102 J106 JU101 JU102 4-20 mA D P A No 1-5 V not field-selectable * 0-10 V B P V No
* Note: Conversion to or from 1-5V must be done at the factory or a Simpson Authorized Service Center.
Note: JU101 and JU102 are hard wire jumpers, and are removed by
cutting them. Resoldering the JU jumpers is not recommended. If this is required, or if a function is to be changed (from volts to current), Simpson recommends returning the Falcon to the factory or an Authorized Service Center. After moving the jumpers to the desired location, put the Falcon back together and install in your panel, or proceed to calibration.
See Chart