Lake Shore F71, F41 Quick Start Manual

Page 1
1 http://www.lakeshore.com/products/Gaussmeters/F71-F41-teslameters/
F71/F41 Teslameter Quick Start Guide
Quick Start Guide
F71 Multi-axis teslameter
F41 Single-axis teslameter
Page 2
2
F71/F41 Teslameter Quick Start Guide
Observe these general safety precautions during all phases of instrument operation, service, and repair. Failure to comply with these precautions or with specific warnings elsewhere in this manual violates safety standards of design, manufacture, and intended instrument use. Lake Shore Cryotronics, Inc. assumes no liability for Customer failure to comply with these requirements.
The F71/F41 teslamter protects the operator and surrounding area from electric shock or burn, mechanical hazards, excessive temperature, and spread of fire from the instrument. Environmental conditions outside of the conditions below may pose a hazard to the operator and surrounding area.
Indoor use
Altitude to 2000 m
-20 °C to 70 °C, <90% relative humidity non-condensing
Overvoltage category II
Pollution degree 2
Ground the Instrument
To minimize shock hazard, the instrument is equipped with a 3-conductor AC power cable. Plug the power cable into an approved 3-contact electrical outlet or use a 3-contact adapter with the grounding wire (green) firmly connected to an electrical ground (safety ground) at the power outlet. The power jack and mating plug of the power cable meet Underwriters Laboratories (UL) and International Electrotechnical Commission (IEC) safety standards.
Ventilation The instrument has ventilation holes in its side covers. Do not block these holes when the instrument is operating.
Do Not Operate in an Explosive Atmosphere
Do not operate the instrument in the presence of flammable gases or fumes. Operation of any electrical instrument in such an environment constitutes a definite safety hazard.
Keep Away from Live Circuits Operating personnel must not remove instrument covers. Refer component replacement and internal adjustments to qualified maintenance personnel. Do not replace components with power cable connected. To avoid injuries, always disconnect power and discharge circuits before touching them.
Do Not Substitute Parts or Modify Instrument Do not install substitute parts or perform any unauthorized modification to the instrument. Return the instrument to an authorized Lake Shore Cryotronics, Inc. representative for service and repair to ensure that safety features are maintained.
Page 3
3 http://www.lakeshore.com/products/Gaussmeters/F71-F41-teslameters/
F71/F41 Teslameter Quick Start Guide
Cleaning Do not submerge instrument. Clean only with a damp cloth and mild detergent. Exterior only.
Desktop Installation When installing the instrument in a desktop environment, ensure it is mounted on a flat, level surface.
Improper Use If the instrument is used in a manner that is not specified by Lake Shore, the safety protections provided by the instrument are no longer guaranteed, and may be impaired.
Child Safety This equipment is not suitable for use in locations where children are likely to be present.
Key specifications
Power requirement 100 V to 240 V (universal input), 50 Hz to 60 Hz, 30 VA
Size 216 mm wide × 87 mm high × 369 mm deep (8.5 in × 3.4 in × 14.5 in), half rack
Weight
3.2 kg (7 lb)
Approval CE mark, FCC, Giteki mark
Operating conditions Rated accuracy: 18 °C to 28 °C, <70% relative humidity non-condensing Reduced accuracy: -20 °C to 70 °C, <90% relative humidity non-condensing
NOTE: Not all specifications are listed. For full specifications, see http://www.
lakeshore.com/products/Gaussmeters/F71­F41-teslameters/pages/Specifications.aspx.
!
Direct current (power line)
Equipment protected throughout by double insulation or reinforces insulation (equivalent to Class II of IEC 536—see Annex H)
CAUTION: High voltages; danger of electric shock; background color: yellow; symbol and outline: black
CAUTION or WARNING: See instrument documentation; background color: yellow; symbol and outline: black
Off (supply)
On (supply)
Frame or chassis terminal
Protective conductor terminal
Earth (ground) terminal
3
Three-phase alternating current (power line)
Alternating or direct current (power line)
Alternating current (power line)
Page 4
4
F71/F41 Teslameter Quick Start Guide
Introduction
This guide provides basic information for getting started with your Lake Shore F71/F41 teslameter. For further documentation and information, see our website.
Items included with the F71/F41 teslameter:
F71/F41 teslameter 6-pin terminal block, two count Line power cord Wi-Fi antenna USB A to USB-Type C™ adapter
Unpacking
1. Inspect all items for both visible and hidden damage that occurred during shipment. If there is visible damage to the contents, contact the shipping company and Lake Shore immediately.
NOTE: Procedures vary with shipping companies. Keep all damaged shipping materials and contents until instructed to either return or discard them.
2. Keep the container and shipping materials until all contents have been accounted for.
3. Check off each item on the packing list as it is unpacked.
Features
The F71/F41 teslameter includes features such as:
DC and RMS measurement Peak to peak measurement Autoranging Variable averaging window Min/Max hold Relative readings Temperature compensation Frequency measurement of field TruZero™ technology to reduce measurement noise
and offset
TiltView™ touchscreen for improved viewing angles Smaller, ultra-thin Hall sensor active areas for
improved accuracy
Multiple probe types to suit your application Fast warm up Isolated input and output relays Analog Hall voltage output 3-year standard warranty High-speed data acquisition Field control option USB, Ethernet and Wi-Fi connectivity
Page 5
5 http://www.lakeshore.com/products/Gaussmeters/F71-F41-teslameters/
F71/F41 Teslameter Quick Start Guide
The front panel consists of:
1. Power button
2. TiltView™ touchscreen
Front panel Rear panel
The rear panel consists of:
1. Option card slot
2. Line input assembly
3. USB Type-C™ interface
4. USB serial communications interface
5. RJ-45 Ethernet interface
6. WLAN antenna connector
7. Digital I/O port
8. Probe input connector
9. BNC analog output connector
q
o i
u
y
t
r
e
w
q
w
Page 6
6
F71/F41 Teslameter Quick Start Guide
Placement
The F71/F41 teslameter is an out-of-the-box benchtop instrument with an adjustable TiltView™ screen for an improved viewing angle. The screen adjusts from a 0° to a 37° viewing angle, whether mounted in a rack or on a bench top.
Rack mounting
The teslameter can be installed into a full rack or half rack mount using the optional Lake Shore rack mount kits (half rack mounting is shown below).
NOTE: Ensure that there is 1 in (25 mm) clearance on both sides of the instrument after rack mounting.
Assemble the blank panel and tray to the rack mount ear before mounting the tray to the instrument
If a non-XIP full rack instrument will be installed above, temporarily install supplied spacers to allow screen to tilt freely
*A third-party shelf may be used, though more space may be required above instrument
Expose threads by removing the feet
using a flat blade
screwdriver
Item Description Qty
1 Rack mount panel 1 2 Rack mount ear 1 3 Half rack mount tray 1 4 Screw, M4 × 5 mm FHM Phillips 6 5 Washer, 1.6 to 1.8 mm THK 2
2U*
1 in (25 mm)
clearance
Standard 19 in (483 mm)
rack width
Page 7
7 http://www.lakeshore.com/products/Gaussmeters/F71-F41-teslameters/
F71/F41 Teslameter Quick Start Guide
Connections and Installation
The teslameter includes a 3-conductor power cord that mates with the IEC 320-C14 line cord receptacle on the back of the teslameter.
WARNING: Always plug the power cord into an easily accessible, properly grounded receptacle to ensure safe instrument operation.
Startup
Connection
Plug in the teslameter using the supplied power cord. The instrument will begin its power-up sequence, after which the Home screen is displayed.
NOTE: The screens you see may differ, depending on your software version. Some screens are scrollable.
Page 8
8
F71/F41 Teslameter Quick Start Guide
Basic operation
Readings
The Home screen displays readings such as magnitude, max hold, relative, XYZ (F71), and frequency. Touch a value to navigate to a detail page containing more information. The Magnitude detail screen for a single axis probe measuring an AC field is shown below.
From the quick access sidebar, you may change parameters such as:
Measurement
Mode: On this screen, choose the mode you wish to use, then click OK.
Averaging Window:
Select a time interval setting, then click OK. A longer averaging window will result in more stable readings, but will respond more slowly to field changes.
Page 9
9 http://www.lakeshore.com/products/Gaussmeters/F71-F41-teslameters/
F71/F41 Teslameter Quick Start Guide
Range: Sets the maximum readable field and
display resolution. The F71/F41 teslameter is equipped with an autoranging feature that will automatically select the appropriate range for the measured field. Select the range you wish to use, then click OK.
NOTE: Autorange is the default setting.
Navigation drawer
Tap the Settings menu to go to the navigation drawer to adjust settings.
Measurement setup: Includes settings for
Measurement mode, Averaging window, Range, Units, and Temperature compensation.
Sensor Information: Provides general
information about the probe.
Analog output: Selects which channel’s analog
signal is routed to the output.
Digital I/O: Use this screen to configure the
functionality of the inputs and outputs.
System settings: Use this screen to adjust
settings, including Wi-Fi, display and sound, firmware updates, and privacy
Page 10
10
F71/F41 Teslameter Quick Start Guide
Measurement modes
DC measurement mode In DC measurement mode, the display shows the measured DC field, the DC max hold, the DC relative measurement.
AC measurement modes In this mode, the display shows the RMS reading as well as the frequency, RMS max hold, and the RMS relative measurement.
The F71/F41 teslameter offers two AC measurement bands to improve overall AC measurement performance:
AC mode: has an operating range from DC to 500 Hz, but
should be limited to frequencies less than 60 Hz for rated accuracies (reports both AC and DC components, and has improved measurement performance)
High frequency mode: 50 Hz to 100 kHz (does not display DC
component, but can connect an oscilloscope to the analog output to view higher frequency signals, allowing high speed pulse capture)
Page 11
11 http://www.lakeshore.com/products/Gaussmeters/F71-F41-teslameters/
F71/F41 Teslameter Quick Start Guide
Max Hold
The maximum and minimum values are displayed on the Home screen. Reset these values by tapping the Reset button. Tap the box to see a detail page where the maximum and minimum can be reset independently.
Relative Measurement
The relative measurement allows you to take field measurements relative to a baseline field. The relative reading is equal to the pres­ent field value minus the baseline. The relative mode applies to the DC reading in DC mode, and to the RMS reading in the AC and high frequency modes.
Touch the Tare button to capture the present reading as a new base­line. Tap the box to see a detail page where the baseline can be set manually.
TIP: This feature can be used to remove the small amount of TruZero™ residual offset if a 0 reading is desired when the probe is placed in a zero gauss chamber. Measure the residual offset with the probe in a zero gauss chamber and set the relative baseline to the average offset value.
Page 12
12
F71/F41 Teslameter Quick Start Guide
Temperature Measurement
The temperature compensation detail screen displays the current temperature reading for the probe tip temperature sensor. Temperature compensation is on by default but can be turned off for trouble­shooting, or set to manual to allow tem­perature inputs that are outside the stan­dard operating range of the probe temperature sensor (0° to 90° C).
Screen Lock
The screen lock feature prevents accidental changes to parameter values. When the screen is locked, some parameter values may be viewed, but not changed, from the front panel.
To lock the screen:
1. Tap the Action icon (three dots in the top right corner of the screen). A drop-down menu appears.
2. Tap Lock. A message indicating that the screen is locked appears in the status bar.
To unlock the screen:
1. Tap the lock icon in the upper right corner of the screen.
2. Tap UNLOCK in the status bar. The screen unlocks.
Page 13
13 http://www.lakeshore.com/products/Gaussmeters/F71-F41-teslameters/
F71/F41 Teslameter Quick Start Guide
Probe accuracy considerations
The user must consider all the possible contributors to the accuracy of the reading. Both the probe and teslameter have accuracy specifications that may impact the actual reading.
Probe readings are dependent on the angle of the sensor (Hall sensor) in relation to the magnetic field. Maximum output occurs when the flux vector is perpendicular to the plane of the sensor. This is the condition that exists during factory calibration. The greater the deviation from orthogonality (from right angles in either of three axes), the larger the error of the reading. For example, a 5° variance on any one axis causes a 0.4% error, a 10° misalignment induces a 1.5% error, etc.
Effect of angular variations on percentage of reading error where percent error = (1 – cos α) 100
Deviation from perpendicular (α)
Error
+B
0% 0°
0.4%
5°
1.5% 10°
3.4% 15°
20°
30°
6.0%
13.4%
29.3%
45°
Page 14
14
F71/F41 Teslameter Quick Start Guide
Computer interface connections
If desired, attach the teslameter to your PC using
Ethernet, USB, or Wi-Fi.
USB Serial port settings
Baud rate 115,200
Data bits 8
Parity None
Stop bits 1
Flow control RTS/CTS
Ethernet/Wi-Fi
IP address DHCP or Manual
Port 8888
Page 15
15 http://www.lakeshore.com/products/Gaussmeters/F71-F41-teslameters/
F71/F41 Teslameter Quick Start Guide
If you are unable to see the teslameter in this view, you may need to install the USB driver that can be found on the Lake Shore Software web page: http://www.lakeshore.com/software/.
Use the SCPI command list (see the user’s manual) and your favorite serial terminal program (such as PuTTY or Termite) to communicate with the teslameter via the remote interface. The USB connection will be listed as a virtual COM port on your PC. If you are not able to connect, make sure you have the correct COM port selected. You can do this by checking DeviceManager:
Page 16
16
F71/F41 Teslameter Quick Start Guide
Rev. 1.2 P/N 119-491 12 November 2018
Contacting Lake Shore
The Lake Shore Service Department is staffed Monday through Friday between the hours of 8:00 a.m. and 5:00 p.m. EST, excluding holidays and company shut down days: http://www.lakeshore.com/Service/
The Lake Shore Forum is also a great place to look for solutions, to post issues, and to share successes:
http://forums.lakeshore.com/
For further documentation and information, please see
https://www.lakeshore.com/products/Gaussmeters/F71-F41-teslameters
Lake Shore Service
Lake Shore Cryotronics Instrument Service Department 575 McCorkle Blvd. Westerville, Ohio USA 43082-8888 Phone: 614-891-2244 Email: [email protected] Web: www.lakeshore.com
Loading...