These notes are intended to draw your attention to risks
which only
YOU
can recognize and avoid or overcome. They
are intended to enhance your own safety consciousness.
We have set the highest quality standards for ourselves and
this unit during development and production. Every unit
meets relevant safety regulations.
The correct unit usage
and proper handling is however solely your responsibility.
The intended workplace should correspond to a laboratory
or pilot plant environment. The user should have an education level which is at least equivalent to a trained laboratory
worker or specialized chemist. The following list should be
seen as an example.
The device may not be operated if there are any doubts
regarding a safe operation due to the outer appearance
(e.g. damages).
A
safe operation of the instrument cannot be guaran-
teed if the user does not comply with this instruction
manual.
Ensure that this manual is always at hand for every unit
operator.
Only use this unit solely for the intended application.
Repairs, alterations or modifications must only be carried out by specialist personnel. Considerable damage
can be caused by improper repairs. The Thermo Haake
service department is at your disposal for repair work.
Do not operate the unit with wet or oily hands.
Do not expose the unit to spray water or immerse it in
water.
Do not clean the unit with solvents (fire risk!)
-
a wet
cloth soaked in household detergent is normally sufficient.
This device is not designed according to the standard
EN 60601-1: 1990 (DIN VDE 0750-1 and IEC 601-1) and
should not be operated in rooms used for medical purposes and/or in the vicinity of patients.
a result of normal
unit functioning -there is a high risk of burns!
The overall temperature of the marked zone (see fig.)
will become higher than
70°C
when the bath tempera-
ture exceeds approx.
150°C.
Please ensure that ade-
quate contact protection is provided.
Do not move the unit from the position where it was set
up during operation or when it is still hot. There is a high
risk of burns!
!
Only use the heat transfer liquids recommended by
Thermo Haake. Please refer to the respective
EC
Safety
Data Sheet.
!
The temperature controlling i.e. immersing of test
tubes, Erlenmeyer flasks or similar objects directly
within the circulator constitutes normal circulator
practise.
We do not know which substances are contained within
these vessels. Many substances are:
inflammable, easily ignited or explosive
hazardousto
health
environmentally unsafe
i.e.:
dangerous
You alone are responsible for the handling of these substances!
Our advice:
If in doubt, consult a safety specialist.
Read the product manufacturer's or supplier's
The following measures were taken for the protection of
the operator:
Protection Class I according to VDE 01 06 TI (IEC 536)
i.e, protection against electric shocks by grounding all
parts which carry the risk of electric contact.
0
The device must only be connected to mains receptacles with a protective ground.
Protection lP 30 according to EN 60529 for all temperature control units
F
and N, i. e. regarding the protection
against accidently touching live parts and damage by
foreign matter, it has been ensured that foreign bodies
with a thickness or diameter of more than 2.5 mm can-
not penetrate.
Protection IP 20 according to
EN
60529 for all cooling
units, i. e. regarding the protection against accidently
touching live parts and damage by foreign matter, it has
been ensured that foreign bodies with a thickness or diameter of more than 12 mm cannot penetrate.
0
NO special precautions were taken against the penetration of water and dust. The device should there-
fore not be used in a dusty atmosphere or in the neigh-
borhood of spray water.
0
Do not insert wires or tools in any of the openings.
!
Complete separation from the mains is required
when:
-
all dangers caused by this device are to
be avoided,
-
cleaning is carried out,
-
repairs or maintenance work is about to be
carried out.
variably adjustable excess temperature protection and independent low liquid level protection which is preset to the
lowest level allow the usage of different heat transfer liquids.
If a safety element is triggered..
.
the cause for the fault is displayed,
the
safety-relevant components
(heating element,
motor and compressor) are switched off immediately
i.e. the safety circuit transfers the unit to a stable, safe
condition,
the temperature of the heat transfer liquid gradually
adjusts to ambient temperature.
5.3
Applications
As open-bath circulator:
For temperature controlling samples within the circula-
tor's own bath.
As heating circulator:
For temperature controlling closed temperature control
circuits such reactors, heat exchangers or similar objects.
Temperature controlling of open vessels using the
built-in combined pressure and suction pump.
5.4
Temperature ranges
Working temperature range:
The temperature range of the circulator without additional heating or cooling sources.
Operating temperature range:
The temperature range of the circulator which can be
reached if additional heating or cooling sources are
used.
Tap water can be used as a cooling source. In this case
the minimum working temperature possible is approx.
Notify carrier (forwarding merchant, railroad) etc.
Compile a damage report.
Before return delivery:
Inform dealer or manufacturer
(Small problems can often be dealt with on the spot).
6.2
Contents
2
Coupling Nuts
(already assembled) 1 Instruction Manual
2
Plug pieces
(already assembled)
2
Hose fittings
for hoses 8 mm
2
Hose fittings
for hoses 12 mm
(E),
(D),
0
(B),
0
(C)
Up
4
Hose clamps,
1 Warranty Card
(please
fill
out and return)
6.3
Ambient conditions according to
indoors, max. 2000 meters above sea level,
ambient temperature
relative humidity max. 80%/31
excess voltage category I I, contamination level
6.4
Resting time after transportation
(only for refrigerated circulators)
As we can unfortunately not guarantee that our refrigerated
circulators are always transported according to our recommendations (i.e. upright), lubrication oil can leak from the
compressor into the cooling circuit.
If the refrigerated circulator is started up whilst still in this
state, the compressor may be damaged to the lack of oil.
Therefore:
Rest the unit for
!
6.5
Ventilation
5
...
40"
C,
24
hours after setting up.
"C
EN
61010
(+
50%/40°C)
2
Keep all ventilation grids (on front and rear) free from
obstruction to ensure unhindered air circulation.
Blocked ventilation grids lead to increased unit
(
=
heating which in turn reduces the cooling capacity
and thus impairs correct functioning.
Thermo Haake circulators and cryostats carry the CE sign
which confirms that they are compatible with the
EU
guideline
89/336/EEC (electromagnetic compatibility). The tests are
carried out according to module
H
(official sheet L380 of the
European Community) as our quality assurance system
is
certified according to DIN
1
IS0
9001.
The specialist basic standards to be applied are
EN
50081
-2
for interference emission and
EN
50082-2 for interference
resistance. The following tests were carried out:
fa according to
EN
50081
EN
5501 1 class B (interference voltage)
EN
55011 class
B
(interference radiation)
EN
50082
EN
61
000-4-2 (discharging static electricity)
ENV
501
40
(electromagnetic
HF
field,
amplitude modulated)
EN
61 000-4-4 (quick transient interference
variable)
ENV
501
41
(high-frequency asymmetrical,
amplitude modulated)
EN
61
000-4-8 (magnetic field with power-
engineered frequency)
There are thus no limitations placed on usage.
A
declaration
of conformity can be supplied with the ordered unit on
request.
Our strict standards regarding minimum operating quality and
the resulting considerable amount of time and money spent
on development and testing reflect our commitment to guarantee the high level of quality of our products even under
extreme electromagnetic conditions. Practice however also
shows that even units which carry the
CE
sign such as monitors or analytical instruments can be affected if their manufacturers accept an interference (e.g. the flimmering of a monitor)
as the minimum operating quality under electromagnetic
compatibility conditions. For this reason we recommend you
to observe a minimum distance of approx.
1
m from such units.
The CE-sign also certifies the conformity with the EU-directive
Socket for external alarm device
Socket for analog control signals (option)
Socket for external Pt100 sensor
RS
232C
Name plate
Pump inlet: back flow from the external object
Pump outlet: pressure to the external object.
Cooling coil fittings for tap water cooling (not for
cooling coil is delivered with the bath vessel);
not in conjunction with refrigerated baths
Mains cable (or cable to the refrigerated bath)
Fuses (not in conjunction with refrigerated baths),
if the fuses are triggered, see chap. 11.4
Control cable to the refrigerated bath
Handle
Drainage nozzle
Opening for temperature control module
Bath opening with bath cover (standard feature)
Ventilation grid (removeable, four mounting points:{)
Fuses
(if
the fuses are triggered, see chap.
11.4)
Socket for cable
30
from temp. control unit FIN
Mains cable
Socket for control cable
return flow from external object
outlet to external object (pressure side)
Hoses are normally used to connect the pump with an external vessel.
If
objects are to be temperature controlled in the
internal bath only, the pump nozzles
A
can be closed with a
,
.
3
@
covering plate D attached with a union nut E (supplied as
L.
standard). However, in order to achieve a better temperature
constancy, it is recommended not to close but to connect the
A
D
E
two nozzles with a short hose with
a
min.
length of
50
cm.
General recommendations concerning the max. allowable
length of hoses cannot be given. It all depends largely on the
size, form and material of the external vessel to
be
tempera-
ture controlled. It should be understood that the length of
a
hose and its diameter combined with the circulating capacity
have a large effect on the temperature control effectiveness.
Whenever possible, the decision should be made in favor of
the wider hose diameter and the vessel to be temperature
controlled should be placed as close as possible to the circu-
lator.
!
High operating temperatures will lead to high tem-
peratures on the hose surface, this is even more so
at the metal
nozzles. In this case:
DO
NOT TOUCH!
!
The required hose material
is
dependent on the
heat transfer liquid used.
!
Hoses must not be folded or bent!
A
wide
radius should
be
used
if
turns have to be
made!
!
Hoses may become brittle after prolonged use or
they may get very soft. They should, therefore, be
checked regularly and exchanged
if
necessary!
!
Secure all hose connections using hose
clamps!
8.2
Selecting Hoses
Thermo Haake circulators will be delivered without any
hoses. Due to the unknown application at the time of shipment and the extremely wide temperature range of the circulators it became impossible to deliver hoses as standard
accessories. Please select the proper hoses from the following table.
Close pump pressure and return port with the closing
pieces and coupling nuts (see chap.
8.1)
or, better yet,
connect the two nozzles with a short hose.
8.5.2
Connection
of
external closed systems
E.g.
instruments with a pressure-tight temperature
jacket or coil or a heat exchanger (external system).
Hose connection:
From the pressure port to the exter-
nal system and then back to the return port.
If
it cannot be avoided that the external object is si-
tuated higher than the circulator, the heat transfer will
only not flow back on the condition that the system is
completely tight and leak-free. To be on the safe side
it may be considered necessary to fit stop cocks to the
inlet and outlet hoses.
8.5.3
Connection of external open systems
Hose connection:
From the pressure port to the exter-
nal bath and then back to the suction port.
With a hose support (optional accessory) which is
employed for baths with a wall thickness up to
26
mm,
both the pressure and suction hose are securely held.
With a clamp in the pressure hose the amount of the cir-
culating liquid is balanced with that of the amount float-
ing back. It is recommended to use a hose with
8
mm
interior-0 as pressure hose and one with
12
mm inter-
ior-0 as suction hose.
The end of the pressure hose in the bath vessel should
be placed in a position where an optimum circulation
within the bath is achieved.
The liquid level of the external bath can
be
adjusted
with the end
of
the suction hose.
The external system and the temperature unit have to
be arranged so that they have the same liquid level in
order to prevent draining by siphoning action. In case
the application requires that both systems have to be
situated at different levels, the two hose lines have to
be closed prior to turning off the temperature unit.
The selection of the proper bath liquid (heat transfer liquid)
influences the capacity of
a
temperature control unit decisively. The technical data with special emphasis on the temperature accuracy was established in accordance with
DIN
58
966
(water at
70°C).
The temperature accuracy will decrease the higher viscosity
of the heat transfer liquid and the lower its heating capacity
IS.
It is difficult to arrive at valid statements which can be applied
as a general rule as the length of the hoses, the volume and
the material of the connected systems have
a
great influence
on this accuracy.
The heating up and the cooling down time of a system to be
temperature controlled can be influenced by the bath liquid
too. Oil, for instance, cuts this time in half when compared to
water.
9.1
Recommended bath liquids
Distilled
Water
Normal tap water leads to calcareous deposits neces-
sitating frequent unit decalcification.
!
Calcium tends to deposit itself on
the
heating element. The heating capacity is reduced and service
life shortened!
Water, of course, can be employed up to 95"C, howev-
er above 80°C water vaporization reaches a level
which necessitates the liquid to be constantly replenis-
hed.
Water
with
Antifreeze
In applications below 5°C the water has to be mixed with an
antifreeze. In doing so, the amount of antifreeze added
should cover a temperature range
10°C
lower than the operating temperature of the particular application. This will prevent the water from gelling (freezing) in the area of the evap-
orating coil the surface area of which
is
much colder than the
working temperature. An excess of antifreeze deteriorates
the temperature accuracy due to its high viscosity.
SIL
I80
...
this heat transfer liquid is suitable for covering nearly the
entire range with just one liquid especially when used with
the cooling units C25, C40 and C41.
Keep level somewhat lower. Oil expands when being
heated. Rule of thumb: 10% volume increase per 100°C heat
increase.
External systems included within the circulating circuit have
to be filled with the same heat transfer liquid in order to avoid
too much liquid being drawn from the internal bath.
#
The bath level should be checked when the preset tem-
perature has been reached!
Quite often closed external systems cannot be prefilled as
suggested.
In this case the internal bath of the unit has to be
ALARM
ji
filled to the max. level. After starting the unit, the pump will
feed the necessary liquid to the external system. Should the
-
LEVEL
demand be higher than the volume difference between high
and low, the low liquid level sensor will be activated and the
Only attach this unit to mains sockets with a grounded earth.
Compare the local mainsvoltage with the specifications writ-
ten on the name plate. Voltage deviations of
+/-
10%
are permissible. The socket must be rated as suitable for the total
power consumption
of
the unit.
Make sure that the temperature controller
F8
or
N8
are
safely connected to the refrigerated units with their
two
cables
30
and
32.
11.2
Checking the liquid circuit
Before switching on, check again to make sure that the pressure and suction ports are either connected with each other
or blocked with covering plugs or alternatively if an external
object is to
The internal control sensor is an immovable fixture. Any
commercially available
Pt100
sensor equipped with 4-wire
technology can be used as the external sensor. For connect-
ing up see chapter
17.
The circulator must be switched
off
and
on
again after inserting an external sensor. The function EXTERNAL is otherwise not selectable. The unit switches automatically back to
the function INTERNAL.
c!l
Select CONTROL, then INT-EXT and choose if you
want to work in the INTERNAL or EXTERNAL mode (or
in the "FOLGEREG" mode).
Three different control speeds are selectable in the EXTER-
NAL mode.
@
Select CONFIG
-..
SETUP-2
-+
DYNAM.EX
The unit is preset to the control speed STANDARD. Overswinging is thus avoided for the most part. The time until the
desired temperature has been reached can however be
quite long. If you would like to shorten this, select MEDIUM
or FAST. When using the setting FAST you should reckon
with quicker control times. The setting
MEDIUM
results in
moderate overswinging and medium control speeds. The
level of overswinging is dependent on a number of factors
such as the volume of the external system. the heat transfer
liquid used, hose length, the working temperature and many
others. No general statements can thus be made at this
point.
"FOLGEREG" mode: temperature control is depending on
the measured temperature at the external sensor.
*To activate this menu the menu key has to be pressed while
switching on the unit until the first display message appears.
13.3
Working with or without cooling
You should decide if the usage of the cooling aggregate (if
fitted) is necessary depending on the desired set temperature.
Select CONTROL and then COOLING.
Switch the cooling aggregate ON or OFF.
The crossed-out star is on when the cooling unit is off.
If the set temperature value is higher than 100°C the cooling unit will not be switched on. Nevertheless, for special
applications, e.
g.
an expected exothermal reaction, it could
be suitable to let the cooling unit run (only partial cooling
capacity of
30%
of the full capacity).
Select CONTROL, then COOLING and then COOL-HT.
Switch to ON if the cooling unit should run (default set-
ting is OFF).
13.4 Setting the desired set temperature
The settings apply to the internal or external control sensors
according to the selected mode INT-EXT (see chapter
13.2).
Select T-SELECT and then SETPOINT. The currently
valid value is displayed in the SETPOINT line. The
SET-NEW line indicates the alteration.
Use the arrow keys to select the digit to alter.
Alter the value between
0 and
9.
Confirm the changes with the ENTER key.
The new value is accepted in the SETPOINT line.
IDENTflashesfor ashort time after the desired temperature has been set: The
FuzzyStar-controller is determining the right parameter. It can happen that heating
or cooling is interrupted during this process and the displayed temperature changes. This is because the control unit requires some time to determine the parameters.
To switch off the identification please see chapter
15.9.
The input is blocked if a temperature setting over I10 port
already exists or the programmer is active. The message
"110
started" or
"PG
started" will be displayed.
13.5 Setting the
RTA
system correction value
The display shows the actual temperature at the internal
or
external control sensor with the selected resolution.
This temperature does not correspond directly to the tem-
perature in the circulator's bath and even less to the temperature in the external connected system.
The temperature difference
is
determined by measuring the
actual current temperature using a suitable measuring device (calibrated or gauged thermometer).
The cut-off point is set with the excess temperature dial with
a rough scale
of
temperature values arranged around it. This
scale, of course, can only serve as an approximate setting
meansfor this cut-off point. However, the cut-off point can be
determined to act exactly if the following procedure is ad-
hered to:
If for instance
a
bath liquid has a fire point of
60°C
the unit
should cut off after reaching 35°C at the latest:
a
First set the desired set value to exactly 35°C.
@
After the circulator has reached this temperature, turn
the excess temperature dial backwards very slowly (to
the left) until the unit cuts off (acoustic signal, alarm is
flashing).
Then set the set temperature to the actual temperature
(<
35°C).
@
Reset the unit via the reset key after the heat transfer
liquid has cooled down somewhat.
-
The unit can now be used for temperatures below
35°C. As soon as
35°C
is reached, it is securely
switched off.
14.1.2
Testing
the
cut-off point
Set the set temperature to a higher value than 35"C, set the
unit to heat up and watch the digital display. The value indicated when the alarm goes on is the real cut-off temperature.
soon as an error occurs in the cooling unit. If the error
can be further specified one of the following messages
appears soon after the first error message has appeared. If the error cannot be specified, the message
"COOLING-RELAIS" keeps visible in the display.
If the compressor of the cooling unit is overloaded the
circulator will be switched to fault status: "COOLING".
Allow the unit to cool down for a few minutes and then
try to start up again. If the fault occurs again
return unit for servicing.
The alarm message "COOL.-VENTI" appears in case
of an overloading of the motor caused by the
compressor
The message "COOL.-RES." appears in case of an
motor overload caused by the compressor 2.
in both cases return unit for servicing.
"
COOL.-V24" appears there is a communation error
If
between the FIN head and the cooling device.
Restart the unit. If this error appears again:
return unit for servicing.
The message "COOL.-COUPL" informs about discon-
nection of the
Check the connection and fuses and restart the unit.
Arrange for constant temperature conditions and
try another identification run.
As this fault only occurs during the identification,
switch off the identification if necessary:
Select CONTROL, then INT-EXT, then FUZZY-ID
and define whether to switch on or off the identification.
If the fault occurs repeatedly:
return unit for servicing.
5
occurs while a program is running:
4
occurs, the identification is considerably dis-
'
ALARM
-
SRAM (SF,
dt):
20.00'C
F09
1
fault
5
If
level of external cooling or heating during identification.
If the fault occurs repeatedly:
Fault
If
a set temperature is selected that is less than 5°C
above the actual temperature, this set value is reached
without identification. In this case the control parameters are taken from the SRAM.
If these values are faulty, they are not accepted by the
Fuzzy control, and fault
=S
occurs, this indicates the influence of a high
Reduce the influence and try another identification run.
Please contact the Thermo Haake product specialists.
9:
System factors in SRAM invalid
9
occurs.
Select a set temperature that is more than
above the actual temperature. This causes the
Fuzzy control to carry out an identification in order
to determine new control parameters.
Potential free contact with the following pin assignment:
Pin
1
=
make contact
Pin
2
=
middle
Pin
3
=
break contact
Alarm relay in
the
circulator:
The relay contacts 2 and 3 are open in case of an alarm
and when the instrument is switched off.
Rating: max.
30
V
max.
0.1
A
17.3
External
Pt100
sensor
A
sensor in four wire technology is necessary. Only
sensors with shielded wires can be used to fulfill the
EMC
requirements. The shielding must be connected
with the housing of the plug and the sensor shaft.
This sensor has to be connected according to the wiring
diagram.
Pin assignment:
Pin
1
=
current
l
+
Pin 2 =
voltage
U
+
Pin
3
=
voltage
U
-
Pin 4 =
current
I
-
The circulator must be switched off and on again after
inserting an external sensor. The function
EXTERNAL
is otherwise not selectable. The unit switches automatically back to the function INTERNAL.
17.4
110
port
(option!)
This port for analog small voltage will be delivered as
an option and is described separately (see appendix).
17.5
Shielded Cables
In order to keep the electromagnetic noise in the instrument within the tolerable limitGt is indispensable to use
only shielded
catles and high quality plug connections.
The complete contact of the shielding within the plugs
Attach
shielding
plane
is of special importance. Insufficient contact may lead
to noise penetration and result in performance errors.
(=IEC 950) should be connected to the interface of the cir-
culator.
18.4 Setting the desired set value
If the desired set value is set via the computer, this means:
If the interface connection is interrupted, the circulator
controls the temperature to the last set value entered
via the interface.
After the mains .supply current has been switched off
and on, the circulator controls the temperature to the
last value entered via its keyboard and not the value
given by the computer.
18.5 Watchdog function
A watchdog can be initialised via the interface, in order to
generate an acoustic alarm when no trigger command has
been received via the interface within the set timeout period.
The circulator then continues operation with the most
recently manually set (via the menu function) setpoint temperature as safety temperature. The most recently received
setpoint from the interface is therewith no longer valid. Thus
the instrument is not switched-off as it would be in the case
of an ordinary disturbance (liquid level,
BATH
overtempera-
ture, etc.). The response delay time of the watchdog can be
changed via the interface. The default setting after switch-on
is 20 seconds.
The watchdog (WD) setting is displayed on the menu path
WD-V24: 0/0020/+000015
I
I
!---current WD remaining response
delay in seconds
1
I---WD response time setpoint in sec.
(0
...
9999 sec)
I---WD status (0 inactive,
1
active)
By reference to the WD remaining response delay time
indication, after the watchdog has responded, it is possible to determine the instant at which communication
failed. This is possible because the residual time is
shown with sign. The value is positive for as long as the
call up first triggered alarm
call up fuzzy control
call up fuzzy error number
Actual temperature
Call up
actual temperature (internal)
actual temperature (external)
Set temperature
(resolution 0,01 "C)
Call up set temperature
Select set temperature
*)
e.g.:
set temperature= 20i0"~
=>
"S-02000"
set temperature= -1 0,5"C
=>
"S
-1 050";
"-"
replaces 1 st digit
RTA (correction factor c)
(resolution 0,01 "C)
Call up RTA (internal)
Call up RTA (external)
Set RTA (internal)
Set RTA (external)
<ox>
=
{+,-);
operational sign
<value>
=
{0000
...
+/-0255)
=>
0°C
...
+/-2,55"C
"blank" and
"+"
can be used as positive signs,
e.g.: c
=
1,23"C
=>
"c-0123" or "c - +0123"
Temperature limit values
(resolution 0,01 "C)
Call up high limit temperature
Call up low limit temperature
Set
high limit temperature
Set low limit temperature
HL
LL
-
-
"
-
"
=
blank
*)
If a new nominal temperature is preset over the digital interface as long as there is another
value preset over the analog I/O port or the built-in programmer the error
"F136-PG110-SOLL-ERROR" will be displayed. The program sequence is interrupted and the desired
set temperature is activated whilst the start temperature is being reached or the program is
set to "pause".
R HL
IN-SP
6
R-LL
1
INSPI7
W-
HL <value>*)
-
w-L~-<va/ue>**)
-
*)
at values with decimal places, put in a point e.g.
f you select a segment for which not all the parameters have been defined, error
message "F126" appears. In this case either select another segment or define the
Darameters.
KRS
I
-
1
Start ramp
If the circulator has reached the start temperature, you can start the ramp with
VV
-
RS.
The command
W
RP "interrupt ramp" is only available while a program is run-
ling, i.e. if the commands
W
AS <no.> "run to start temperature" or W RS "start
.ampen have been entered bef6e.lf this is not the case
"F127"
appears when
~ntering the command
W
-
RP "interrupt ramp".
!
If a ramp segment runs out, the next segment does not start automatically.
Stop ramp
Interrupt ramp
N
RB
N-RP
-
To start the next ramp segment, once again enter the following commands:
fou can interrupt the program with W-RP. Then the segment time is hold and the
nomentanous temperature is maintained. The interrupted program can either be
:ontinued with W RS or another segment can be selected with
W
-
SNxx.
Fault
F127
=
"PAUSE ERROR"
-
-
Run to start temperature
(segment no.: 1-6)
Start ramp
Fault
F001
=
"COMMAND UNKNOWN
"
If a command is entered that is not defined in any set of commands, error message F001 appears. In this case enter one of the commands listed above.
The single ramps can be repeated cyclically:
Call up for information:
R-R I
-
remaining ramp time and
sensor value.
R RZ
I
-
-
W - RZ
-
<PgNo.> - <number>
-
R - XR
1
-
(Read status ramp continuation
Answer
,,
XR-< program no.
>-<
segment no.
>-<
segment remain time
>-
<
setpoint step
>-<
actual ramp setpoint
>-<
segment end temperature
>
,,
I
Call up number of cycles
Set number of cycles
-
" - "
=
blank
64
Read remaining ramp time and
external sensor value.
control parameters are determined. Then the heating is activated to reach the start temperature of segment no.
(5)
below the end temperature, when the program starts seg-
ment no.
long as the momentary set temperature is higher than the
actual temperature at the beginning of the segment the unit
heats until the two temperature curves intersect.
(6)
decrease given in segment no.
without cooling aggregate. Thus the actual temperature
decrease is slower than the set temperature.
0
15 20 25
After starting the program the heating is switched off and
At the end of the segment the actual temperature is still
4.
A negative temperature gradient is available. As
At this point the heating is inactivated. The temperature
The times of segments no. 3 and no. 4 should be
extended so that the circulator can keep the given
values.
The stainless steel surfaces of the bath vessel and of the
housing may after some time show spots and become tarnished. Normal stainless steel cleaners as they are used in
the kitchen can be used. The bath vessel and built-in components should occasionally (at least every time the bath liquid
is changed) be cleaned using a household cleaner.
based cleaners have proved to be suitable used according
to the manufacturers recommendations.
0=f
Do not use scouring powder!
The inside of the bath vessel must be kept clean in order to
ensure a long service life. Substances containing acidic or
alkaline substances and metal shavings should be removed
quickly as they could harm the surfaces causing corrosion.
If corrosion (e.g. small rust marks) should occur in spite of
this, cleaning with stainless steel caustic agents has proved
to be suitable. These substances should be applied accord-
ing to the manufacturers recommendations.
Vinegar-
20.1 Cleaning the fins of the liquefier
In order to maintain the cooling capacity of the unit, cleaning
has to be done two to four times per year, depending on the
grade of soiling.
Switch off the unit and pull out the mains plug.
!
Loosen the ventilation grid on the front: Rotate the
90"
mounting screws
Clean fins with brush or similar tool.
Replace grid and push screws back in (do not rotate
screws).
20.2 Discarding the unit:
One day the life span of your cooling unit will end.
Therefore:
Unscrew the cover caps of the sockets at the rear
instrument panel.
r
-A&-
-
@
Remove screws at the rear panel
(4
screws).
@-
Remove rear panel.
Bend the hood slightly apart and slide off to the back.
@
Separate the socket of the It0 board at the rated break
point. Insert
I10
board as shown in
Fig.
1
.
(Insert PCB up to the stop of the pin strip)
Fasten socket on housing with a nut (watch torsion
protection).
Slide hood back on.
Fig.
1
Check grounding connections.
@
Screw the rear panel back on.
MAIN
MENU
!
For connection with the voltage
or
mains supply a
shielded cable has to be used.
Should you notice any jumps of the set temperature values
on the display the shielding is insufficient. Please shield as
described in the manual (chapter
17.5).
If
the problem persists (e.g. in case of a longer cable) the cable should perhaps
be embedded in a tube.
In the other case select 0...10 V and the temperature
range O...lOO°C or
-1
00
...
400°C in the
U-INPUT.
On the display the following is shown (from top to bottom):
Upper line U-INPUT the momentarily saved mode
(e.g. U-INPUT-I
:
-1
...
3V) with the channel (e.g. 1) and
the voltage range (e.g. -1
...
3V);
Lower line U-INPUT the momentarily marked mode
(e.g. U-INPUT-O:*O
...
IOV) with the channel (e.g. 0) and
the voltage range (e.g. O...IOV);
Below: the available modes (e.g. 0...1 OO°C, 10 mVI0C,
-1 00
...
400°C).
With the arrow keys you can select a mode. The
respective voltage range and channel are shown in the
lower line U-INPUT.
If
you confirm this selection (i.e.
this channel) with ENTER, the information is taken over
to the upper line U-INPUT.
I
!
1-0 STARTED
!
I
Press the menu key to get to the INPUT menu. Here
with START you can activate the defined voltage val-
ues. From now on the circulator will receive its set values via the pre-set voltage values.
For a short time in the display appears
"!
1-0
STARTED!"
together with the set values (e.g.
O...
1
OV
-+
0
...
100°C).
(
CONFII-OIANAIINISELECT
1
U-INPUT-I: -1 ..3V
!
1-0 STARTED
!
piiZZ-1
10 rnVI"C
!
If a channel is momentarily active, it is not possible
to save a new setting: while you press the ENTER
key,
"!
1-0
STARTED
!"
is shown on the display.
To save a new setting first stop the operation of the
110 plug with STOP in the INPUT or OUTPUT menu.
I
CONFII-OIANALOG
1
I-INPUT I-OUTPUT
23.3.2
Configuration of current input
If
you want to give a set value with an alalog current by
means of an external device, select I-INPUT (current
input) in the ANALOG menu and then SELECTION in
the INPUT menu to set the values for the current input.
When the SELECT menu appears two different modes
can be chosen: normal resolution (-1 00
...
400°C) and
high resolution (0
...
100°C). So with the current ranges
0...20 mA or 4...20 mA four combinations (channels 4
to
Select the temperature range with the arrow keys and
cnoose 4...20mA or 0...20mA in the lower line I-INPUT.
Press the menu key to get back to the
INPUT
menu.
Here with START you can activate the selected values
to start the input via the I/O port. From now on the circu-
lator will receive its set values via the pre-set current.
For a short time
"!
1-0
STARTED !"is shown on the display
with the selected values (e.g.: 0...20
mA,
-1
00
...
400°C).
!
"!
1-0
STARTED
!"
is shown on the display also if
you
try to start the programmer (only
F8
and
N8)
while
the
I10 plug is active.
To start the programmer first stop the operation of
the
I10
plug
with
STOP
in
the
INPUT
or
OUTPUT
menu.
23.3.3
Adjusting the input values
If you give a set value by means of an external device
via the I/O plug, there can be a difference between the
set temperature shown on the display of the circulator
and the set value on the external device.
This difference depends on tolerances in the
110 board
and the external device.
By setting an offset value, the value (i.e. a voltage or
a
current, respectively) shown on the external device can
be adjusted to the temperature shown on the display of
the circulator.
E.g.: voltage input and the normal resolution mode
(0
...
10 V =
O...lOO°C; channel
1)
is selected. On the
external device 8.0 V is set to give a set temperature
of 80°C.
Due to some tolerance, the set temperature on the dis-
play of the circulator is not
80.00°C but
81
.40°C. (The
set temperature is always in the upper right corner of
the display.)
This difference can be adjusted with an offset of
-1
.40°C.
To set the offset get back to the NWT menu with the
In the OFFSET menu you can set the offset value (i.e.
the difference between the two temperatures) with the
arrow keys
t
and
+.
After this, the value on the external device and the set
temperature on the display of the circulator are
adjusted.
Confirm with ENTER to save the offset value.
On the display the voltage or current range of the
selected mode is shown
(e.g. 0...1 OV, O/lOO°C). For
each channei an offset value can be saved.
23.3.4 Configuration of voltage output
If you want to monitor an actual value (i.e. an actual
temperature) with an alalog voltage signal
by
means of
an external device, select U-OUTPUT (voltage output)
in the ANALOG menu and then SELECTION in the
OUTPUT menu to set the values for the voltage output.
The configuration of U-OUTPUT is similar to the one of
U-INPUT. You can choose between the same modes:
a)
0...10 V
=
O...lOO°C (channel 0)
b)
-1
... 3 V =10 mV/OC (-100
...
300°C)
(channel 1)
c) 0...10 V =-1 00
...
400°C (channel 2)
The display shows OUTPUT in stead of INPUT.
23.3.5 Configuration of current output
If
you want to monitor an actual value (i.e. an actual
temperature) with an alalog current signal by means of
an external device, select I-OUTPUT (current output)
in the ANALOG menu and then SELECTION
in the
OUTPUT menu to set the values for the current output.
The configuration of I-OUTPUT is similar to the one of
I-INPUT. You can choose between the same combina-
tions of output current ranges and temperature ranges