Felix Storch F-950 Operation Manual

Operations Manual
F-950
Three Gas Analyzer
DECLARATION OF CONFORMITY
Manufacturer:
CID Bio Science, Inc. Felix Instruments – Applied Food Science 1554 NE 3rd Ave Camas, WA 98607
Declares that the CE-marked Product:
Product Models (s):
Model F-950
Complies With:
89/336/EEC Electromagnetic Compatibility Directive 73/23/EEC Low Voltage Directive
Compliance Standards:
EN 55027 RF Emissions Information
Technology Equipment
EN 50082-1 EMC Immunity Standard
EN 60950 Safety of Information Technology Equipment Including Electrical Business Equipment
October 23, 2014 Leonard Felix President
Table of Contents
Introduction ................................................................................. 1
Features .................................................................................... 3
Specifications ........................................................................... 4
Unpacking the F-950 .................................................................... 5
Operating Instructions ................................................................. 6
Loading the Battery .................................................................. 6
Basic Operation ........................................................................ 8
Measurement Modes ............................................................... 9
Interfering Gases ........................................................................11
PolarCept ................................................................................12
Measure .....................................................................................16
Continuous Mode ...................................................................16
Controls ..............................................................................17
Trigger Mode ..........................................................................18
Fixed Volume Mode ...............................................................20
Auto-Escape Feature ..............................................................24
Setup ..........................................................................................25
Setup Mode ............................................................................25
Setup Calibration ....................................................................26
Setup Set Zero ........................................................................27
Setup Date and Time ..............................................................29
Setup RH Conversion ..............................................................30
Setup GPS ...............................................................................31
Setup Language ......................................................................31
Setup Factory Settings............................................................31
File .............................................................................................. 32
File Select ............................................................................... 33
File New ................................................................................. 34
File Delete .............................................................................. 34
File Review ............................................................................. 35
Data Files on the Computer ................................................... 36
Firmware Update ....................................................................... 38
Installing the F-950 Driver ..................................................... 38
Updating the Firmware .......................................................... 40
F-950 Menu System Diagram ..................................................... 45
Appendix I: F-950 Controller Software and Sensor Calibration . 46
Appendix II: Sampling from a Jar ............................................... 56
F-950 Production Test Check Sheet ........................................... 58
FELIX INSTRUMENTS HARDWARE WARRANTY .......................... 60
Warranty Registration Card ....................................................... 63
F-950 Instruction Manual rev. 10/17/16
1554 NE 3rd Ave, Camas, WA 98607, USA 1 Phone: (360) 833-8835
sales@felixinstruments.com www.felixinstruments.com
Introduction
Congratulations on the purchase of your new F-950 Three Gas Analyzer!
The new F-950 Three Gas Analyzer measures 3 critical gases: ethylene (C2H4), carbon dioxide (CO2) and oxygen (O2), to maintain optimum produce quality at every phase.
The F-950 is used to:
Inspect storage and packing environments Verify ethylene mitigation efficacy Optimize ripening storage atmosphere conditions Quality assure MAP (Modified Atmosphere Packaging)
for ethylene-sensitive products
Ethylene affects ripening, aging, and spoilage in produce. The F­950 measures levels of ethylene, CO2 and O2 in the atmosphere, and can be scaled to many environments, from cold storage to warehouse to transportation container.
Simple to operate and weighing less than a kilogram, the F-950 uses an electrochemical cell to measure ethylene between 0­200 ppm in air. It records date, time, relative humidity, temperature, and coming soon GPS location. The F-950 is ideal for measuring ethylene production across a wide range of fruit or floral types and is especially suitable for managing ethylene emissions from perishable cut produce products.
F-950 Instruction Manual rev. 10/17/16
2 1554 NE 3
rd
Ave, Camas, WA 98607, USA
Phone: (360) 833-8835
sales@felixinstruments.com
www.felixinstruments.com
The F-950 is ideal for Modified Atmosphere Packaging (MAP) applications. Most MAP gas analyzers are only equipped to measure CO2 and/or O2 concentrations. Quality assurance for perishable cut produce demands a more comprehensive analysis. The F-950 makes it possible to respond to the change in ripening gases to ensure longer shelf-life and higher quality commodities.
We hope you enjoy using your F-950 Three Gas Analyzer.
F-950 Instruction Manual rev. 10/17/16
1554 NE 3rd Ave, Camas, WA 98607, USA 3 Phone: (360) 833-8835
sales@felixinstruments.com www.felixinstruments.com
Features
Portable, lightweight and easy to operate.
Rapid response time, with data points saved every second.
Repeatable, precise measurements.
Three versatile modes of operation
Measure in Continuous, Fixed Volume, or Trigger mode depending on
your application.
True sunlight readable transflective display.
The contrast of the display increases under brighter sunlight.
Removable, re-chargeable standard sized batteries.
Included stand-alone battery charger enables charging one battery
set while using another. Two sets of batteries are included with the
device. Additional button-top 19670 (or protected 18650) batteries
can be purchased from your preferred battery vendor.
F-950 Instruction Manual rev. 10/17/16
4 1554 NE 3
rd
Ave, Camas, WA 98607, USA
Phone: (360) 833-8835
sales@felixinstruments.com
www.felixinstruments.com
Specifications
F-950 Specifications
Air Sampling Rate
70 mL/min
Measuring Rate
Automated; 1 sec intervals
Data Storage
Removable 4 GB SD card
Display
Sunlight visible transflective LCD
Operating environment
0°C - 45°C (0-90% humidity non­condensing)
Battery Capacity
7 hours
Dimensions
18 x 13.5 x 5.5 cm
Weight
0.95 Kg
Enclosure
Powder coated aluminum
Warm-up time
< 3 minutes
Sensors
Ethylene Sensor
Electrochemical
Nominal Range
0 – 200 ppm
Lower Detection Limit
0.5 ppm
Resolution
0.1 ppm
Accuracy
5% ±0.5 ppm Continuous mode ±10% Fixed Volume mode
Carbon Dioxide Sensor
Infrared Sensor, Pyroelectric detector
Nominal Range
0 – 20% (200,000 ppm)
Full scale resolution
0.01%
Accuracy
±5% Continuous mode ±10% Fixed Volume mode
Oxygen Sensor
Electrochemical
Nominal Range
0-100%
Resolution
0.1%
Accuracy
±5% Continuous mode ±10% Fixed Volume mode
F-950 Instruction Manual rev. 10/17/16
1554 NE 3rd Ave, Camas, WA 98607, USA 5 Phone: (360) 833-8835
sales@felixinstruments.com www.felixinstruments.com
Unpacking the F-950
The F-950 arrives with a hard-sided carrying case, two sets of batteries and a charger, a removable 4 GB SD card, and several accessory parts. The unit comes with an external conditioning chamber and potassium permanganate (KMnO4), used to scrub the air entering the system. A sampling port with needle is included for fixed volume samples from packaging and an external PolarCept filter for reducing interfering gases. The sampling port is pictured below, connected to the intake.
F-950 Instruction Manual rev. 10/17/16
6 1554 NE 3
rd
Ave, Camas, WA 98607, USA
Phone: (360) 833-8835
sales@felixinstruments.com
www.felixinstruments.com
Operating Instructions
WARNING: Do not store the F-950 without batteries! Charged
batteries must be present in the instrument to maintain the accuracy of the sensors, even when the unit is powered off.
If the batteries of the F-950 discharge during storage, replace with charged batteries and allow the instrument to stabilize 48 hours before use. There is a small internal battery to maintain the bias voltage for the ethylene electrochemical sensor. This small internal battery will last for 1 day without the main batteries before sensor sensitivity is affected by losing its bias voltage. The calibration parameter data is intact with or without batteries.
Fully charged main batteries allow for storage time of over 1 year. For long-term storage, consider attaching the potassium permanganate (KMnO4) external conditioning chamber to the inlet and outlet of the F-950 (see image below).
Aged batteries or batteries that started out with less charge will reduce the storage time available. The Li-ion batteries have little self-discharge and a life of about 3 years.
Loading the Battery
The F-950 uses 18650 Li-ion 3.7V 3100mAh rechargeable batteries. For longer lifespan, charge the batteries at 0.25A. For a faster charge, charge at 1A. The batteries must be removed from the F-950 to be charged.
F-950 Instruction Manual rev. 10/17/16
1554 NE 3rd Ave, Camas, WA 98607, USA 7 Phone: (360) 833-8835
sales@felixinstruments.com www.felixinstruments.com
To remove the batteries, twist the battery compartment cap, located on the bottom of the case. The cap can be twisted with fingers or a screw-driver to tighten or loosen. Take care when removing batteries, as the cap is spring loaded. Both batteries should be inserted into the unit positive (+) side first (towards intake or top). The battery compartment and exhaust port are pictured below.
F-950 Instruction Manual rev. 10/17/16
8 1554 NE 3
rd
Ave, Camas, WA 98607, USA
Phone: (360) 833-8835
sales@felixinstruments.com
www.felixinstruments.com
Basic Operation
To turn the instrument on, press the green power button. The Felix Instruments logo will flash, followed by the main menu. The top of the display reads Felix Instruments, and the current version of firmware the unit is running is displayed in the lower left hand corner. The battery meter is listed on the lower right hand side of the display.
For information on the latest firmware version, please visit the F-950 support webpage (www.felixinstruments.com/support/f-
950-support).
The main menu displays the following options: Measure, Setup, and File. If the green power button is pressed, the display will prompt for confirmation before shutting off the F-950.
To scroll between menu items in the list, use the Up and Down arrows. To select an option from the menu list, use the Right arrow. To exit, use the Left arrow.
F-950 Instruction Manual rev. 10/17/16
1554 NE 3rd Ave, Camas, WA 98607, USA 9 Phone: (360) 833-8835
sales@felixinstruments.com www.felixinstruments.com
Measurement Modes
Three measurement modes are present on the F-950: continuous, trigger and fixed volume. To change measurement mode go to Setup > Mode from the main menu (page 25).
In fixed volume and trigger mode, the instrument reports values in terms of VOC (volatile organic compounds). VOC’s are aromatic compounds like ethylene, esters and alcohols. Their production increase with produce aging and spoilage. The F-950 ethylene sensor cannot distinguish ethylene from many other VOC’s. PolarCept was designed to optimize and improve the accuracy of ethylene measurements. PolarCept is an external water filter, which reduces the signal from interfering gases reaching the ethylene sensor. PolarCept is recommended in Fixed Volume and Trigger mode. PolarCept dramatically reduces the amount of interfering gases like alcohols and esters that are present in a sample, allowing ethylene to pass through to the sensors. For more information on PolarCept, see page 11.
Continuous measurement mode measures the air entering through the input of the instrument. The controls default to
Loop “Open” and Pump “On”. Continuous mode can be used
with or without the sample port attached to the front of the instrument. The data is saved every one second in continuous mode.
Trigger measurement mode begins with the valve closed and the pump off. To start taking a measurement, press the square start button on the key-strip of the F-950 to initiate a new measurement. The pump will run and air will enter the F-950
F-950 Instruction Manual rev. 10/17/16
10 1554 NE 3
rd
Ave, Camas, WA 98607, USA
Phone: (360) 833-8835
sales@felixinstruments.com
www.felixinstruments.com
until a stable reading has been reached for all sensors, or until the user stops the measurement by pressing the start button
again. The final values will be displayed under the word “last” in
the graph and saved to the SD card. The pump will then turn off until the user initiates a new measurement.
Trigger mode can be used with or without the sample port attached to the front of the instrument. The instrument will detect and report total VOC’s present in the sample.
Trigger mode should be used in place of continuous mode if expecting high ethylene concentrations. Trigger mode allows a
purge of the internal volume of gas in the instrument between samples and will help protect the ethylene sensor from over­saturation.
Fixed Volume measurement mode draws in a sample of a small, set volume of gas (default 15 mL). The sampling port with the needle attachment is required for Fixed Volume mode and MAP applications. The instrument will detect and report total VOC’s present in the fixed volume sample.
F-950 Instruction Manual rev. 10/17/16
1554 NE 3rd Ave, Camas, WA 98607, USA 11 Phone: (360) 833-8835
sales@felixinstruments.com www.felixinstruments.com
Interfering Gases
No analytical method is completely specific. Gases present in
the environment, other than the “target” gas of a
measurement, may affect instrument response. Interferences are not necessarily linear, and may also exhibit time dependent characteristics. The following table details the approximate cross-sensitivity of the ethylene sensor to several common interfering gases. A 20% cross sensitivity means that 100 ppm of the interfering gas would read the same as 20 ppm of the target gas (ethylene). Therefore, 20 ppm H2S should read <40 ppm on the F-950: 10 ppm NO2 should read <5 ppm: 10 ppm Cl2 should read -0.1 ppm.
Gas
Interfering gas (ppm)
Cross-Sensitivity (%)
H2S
20 ppm
< 200
NO2
10 ppm
< 50
Cl2
10 ppm
< -1
NO
50 ppm
< 80
SO2
20 ppm
< 50
CO
400 ppm
< 30
H2
400 ppm
< 0.5
C2H4
80 ppm
< 100
NH3
25 ppm
< 0.1
HCHO
4 ppm
90
CO2
50000 ppm
< 0.1
Ripening fruit emit a complex mixture of hydrocarbons, including ethylene. Oxidation of these other gases in the electrochemical sensor cannot be readily distinguished from ethylene. This causes the ethylene value to be falsely high in the presence of interfering gas.
F-950 Instruction Manual rev. 10/17/16
12 1554 NE 3
rd
Ave, Camas, WA 98607, USA
Phone: (360) 833-8835
sales@felixinstruments.com
www.felixinstruments.com
Felix Instruments has tested a method to absorb some of the competing gases and provide better ethylene measurements. This method, PolarCept, uses distilled water in an external chamber and has been shown to filter out some polar hydrocarbons and alcohols to produce less interference.
PolarCept
It is recommended to use the external PolarCept filter when measuring a mixture of gas (such as when sampling fruit) or interfering gases may be reported by the instrument. The external filter can be used with any of the three measurement modes, and is recommended for Fixed Volume and Trigger Modes. PolarCept should only be used with 1.5 mL of distilled
or deionized water.
F-950 Instruction Manual rev. 10/17/16
1554 NE 3rd Ave, Camas, WA 98607, USA 13 Phone: (360) 833-8835
sales@felixinstruments.com www.felixinstruments.com
The PolarCept filter consists of a plastic molded part, hydrophobic filter and O-ring. It is also used with the “fixed volume tubing” and sample probe needle. A small plastic syringe is used to fill and empty PolarCept. Once the hydrophobic filter and O-ring are in place, it should be very difficult to remove it, creating a leak-proof seal. Additional hydrophobic filters are included as replacements, when the filter is soaked with water or damaged during removal.
Eventually the water in the PolarCept filter will become saturated with trapped interfering gases and should be replaced with fresh distilled water. Saturation rates will depend on the measurement mode and amount of interfering gases present in the sample environment. The table below shows example saturation times when measuring headspace of bananas (with a maturity index of 5) in Continuous Mode. This sample contains various mixed hydrocarbons, ethylene and VOCs. The total VOC in ppm listed is the signal reported by the C2H4 sensor in ppm.
VOC concentration
PolarCept saturation (min)
3 ppm
20
100 ppm
1
To fill the PolarCept filter, attach an empty syringe to the hydrophobic filter. The plunger of the syringe should be completely depressed. Lower PolarCept over a cup of distilled water and draw in 1.5 mL with the syringe. Attach the sample needle and fixed volume tubing to stop leaks. To empty PolarCept, re-attach the syringe and push the water out the sample needle end.
F-950 Instruction Manual rev. 10/17/16
14 1554 NE 3
rd
Ave, Camas, WA 98607, USA
Phone: (360) 833-8835
sales@felixinstruments.com
www.felixinstruments.com
To properly use PolarCept, keep the sample needle pointed downwards while measuring, as seen in the example below on
the left. The water in the filter should “bubble” as the gas
sample is pulled through it, causing some of the interfering gases to be trapped.
F-950 Instruction Manual rev. 10/17/16
1554 NE 3rd Ave, Camas, WA 98607, USA 15 Phone: (360) 833-8835
sales@felixinstruments.com www.felixinstruments.com
See the following graph of measured ethylene concentration by the F-950 with and without using PolarCept, measuring the same headspace of bananas. After about 20 minutes, the water becomes saturated, and the signal beings to rise with PolarCept.
0
0.5
1
1.5
2
2.5
3
3.5
0 10 20 30 40 50 60 70
Ethylene (ppm)
Time (min)
F-950 PolarCept
No Polarcept Polarcept
F-950 Instruction Manual rev. 10/17/16
16 1554 NE 3
rd
Ave, Camas, WA 98607, USA
Phone: (360) 833-8835
sales@felixinstruments.com
www.felixinstruments.com
Measure
From the main menu, press the right arrow when the word Measure is highlighted, to enter the measurement display screen. All measurement variables are saved to the SD card every 1 second in continuous mode. When the SD card is not present, the data will not be saved.
If sampling very high concentrations followed by very low concentrations, allow the instrument time to purge internal gas for the most accurate measurements.
Continuous Mode
When in continuous mode, a graph of the concentration of each gas can be viewed over time. The default graph shown is the ethylene (C2H4) concentration in ppm. To view the graph of the other gases, simply use the up and down arrows to scroll through them. The current gas being graphed is shown on the top of the screen with the concentration in large font as shown below.
F-950 Instruction Manual rev. 10/17/16
1554 NE 3rd Ave, Camas, WA 98607, USA 17 Phone: (360) 833-8835
sales@felixinstruments.com www.felixinstruments.com
The x-axis of the graph is time. The y-axis of the graph displays the range of the concentration (in ppm for ethylene and % for CO2 and O2), and the dynamic range is labeled at the top. The y­axis scale is set by the highest value shown in the buffer. This range will scale vertically, dependent on the highest concentration of gas measured. If the concentration is small, the dynamic range will reflect this.
The graph begins on the left side and moves towards the right as more data points are added. Once the line reaches the right side of the display, the data will begin moving towards the far left, keeping the current time at the far right. The total measurement time is displayed below the graph.
Current Oxygen (% O2), Carbon Dioxide (% CO2) are listed on the right side of the graph when ethylene is being graphed. This data changes slightly when displaying the graph of other gas concentrations. The flowrate (mL/m) is displayed at the bottom. If Relative Humidity (% RH) and Temperature (degrees Celsius) are not seen on the display, go to Setup>RH Conversion and enter the actual temperature of the measurement gas (see page
30).
Controls
From the measurement display screen, press the right arrow to view and change the controls. Access to the controls is only available in Continuous and Trigger measurement modes.
The internal volume inside of the instrument, or “Loop” can be
Open or Closed. When the loop is open, the instrument is sampling air from the input at the top of the instrument and exhausting air through the output at the bottom of the instrument. When the loop is closed, air is circulated internally
F-950 Instruction Manual rev. 10/17/16
18 1554 NE 3
rd
Ave, Camas, WA 98607, USA
Phone: (360) 833-8835
sales@felixinstruments.com
www.felixinstruments.com
through the tubes and valves to the input and output are closed.
From the control screen, the Pump can be turned On or Off. When the pump is on, the unit will flow at 67 mL per minute, which is the default flow rate. When the pump is off the flow will drop to 0 mL per minute.
Trigger Mode
A measurement in trigger mode begins with the valves closed and the pump off. To begin, press the square start button as prompted, to measure. Use the longer probe- the Trigger Mode Probe- (seen below) to take Trigger Mode measurements. This will turn on the pump and the ethylene sensor will stabilize before drawing in sample gas. Trigger mode will take longer to stabilize if the concentration is in a different range from the last sample (30 sec-1.5 min).
F-950 Instruction Manual rev. 10/17/16
1554 NE 3rd Ave, Camas, WA 98607, USA 19 Phone: (360) 833-8835
sales@felixinstruments.com www.felixinstruments.com
Once adequate sample has been collected, press the square button to stop the measurement and view the results. The maximum sample volume in this mode is 41 mL. The results will be saved to the SDcard.
Press the square start button to begin another measurement. Press the Left arrow to exit to the main menu.
F-950 Instruction Manual rev. 10/17/16
20 1554 NE 3
rd
Ave, Camas, WA 98607, USA
Phone: (360) 833-8835
sales@felixinstruments.com
www.felixinstruments.com
Fixed Volume Mode
To begin, the unit should be disconnected from the sample of interest while the sensors purge, stabilize and equilibrate to the ambient environment. The instrument will purge for 5 seconds to remove the previous sample from the internal tubing of the instrument. If additional time is required, press the Right arrow to start the purge cycle again. Once stable, the instrument will prompt the user to connect the sample and press start to begin measuring.
F-950 Instruction Manual rev. 10/17/16
1554 NE 3rd Ave, Camas, WA 98607, USA 21 Phone: (360) 833-8835
sales@felixinstruments.com www.felixinstruments.com
Assemble the probe by connecting the “Fixed Volume Needle Probe” black tubing to the intake of the F-950. Next, twist on a
hydrophobic filter to the end of the tubing. The filter will prevent any moisture or debris from being sucked into the instrument. Finally, attach a sterile needle to the filter. Depending on the application, needles can be re-used.
F-950 Instruction Manual rev. 10/17/16
22 1554 NE 3
rd
Ave, Camas, WA 98607, USA
Phone: (360) 833-8835
sales@felixinstruments.com
www.felixinstruments.com
After the start button has been pressed, the pump will run until the sample volume (default: 15 mL) has been drawn into the F­950, and then the loop will close while the sample is analyzed. Maximum sample volume in fixed volume mode is 18.5 mL. Minimum sample volume is 10 mL for 5-10% accuracy.
When the loop is closed the F-950 system should be disconnected from the sample. The unit will read “Processing… after the fixed volume is drawn inside the unit. Once processing is seen on the display, disconnect from the sample while the sensors analyze the fixed volume sample and provide the results.
F-950 Instruction Manual rev. 10/17/16
1554 NE 3rd Ave, Camas, WA 98607, USA 23 Phone: (360) 833-8835
sales@felixinstruments.com www.felixinstruments.com
The final result is then displayed on a new screen. The values listed are the concentrations of C2H4, CO2, and O2 present in the sample. To the right of the concentrations, are the relative humidity, temperature, volume of sample, and duration to collect and analyze sample.
F-950 Instruction Manual rev. 10/17/16
24 1554 NE 3
rd
Ave, Camas, WA 98607, USA
Phone: (360) 833-8835
sales@felixinstruments.com
www.felixinstruments.com
Press the start button to return to the measurement screen. Connect to the next sample and press the start button to begin another measurement.
Auto-Escape Feature
An upper limit auto-escape feature acts as a safety feature of the unit. This safety feature is always on. If the sensor detects over 200 ppm ethylene, the F-950 will auto-escape from monitor mode to prevent poisoning the electrode in the ethylene sensor. This will stop the measurement.
If the sensor becomes poisoned, it will continually auto-escape with the error message "sensor out of bounds". If this message appears, connect the external conditioning tube filled completely with potassium permanganate (KMnO4) to the inlet/outlet of the F-950. Allow time with the instrument powered on (Continuous Mode) to remove the high concentration of ethylene in the instrument.
F-950 Instruction Manual rev. 10/17/16
1554 NE 3rd Ave, Camas, WA 98607, USA 25 Phone: (360) 833-8835
sales@felixinstruments.com www.felixinstruments.com
Setup
Use the right arrow to enter the Setup Menu from the Main Menu screen.
The F-950 has a number of utility functions that allow the user to manage the instrument’s capabilities. Five setup functions are available: Mode, Calibration, Set Zero, Date & Time and RH Conversion, GPS and Factory Setup.
Setup Mode
Setup Mode allows the user to change the measurement mode of the instrument between Continuous, Trigger, and Fixed Volume mode. To change the mode, highlight Measure and use the right arrow to highlight the mode. Then use the Up/Down arrows to cycle through the mode options. Use the left arrow to highlight Measure once the correct mode has been selected. Then press the left arrow again to exit to the main menu.
F-950 Instruction Manual rev. 10/17/16
26 1554 NE 3
rd
Ave, Camas, WA 98607, USA
Phone: (360) 833-8835
sales@felixinstruments.com
www.felixinstruments.com
USB mode should be set to Mass Storage. Virtual COM is the other mode, and should only be used to update the firmware if instructed by a Felix Instruments technician.
Setup Calibration
Setup Calibration lists the C2H4 Offset, C2H4 Gain, Internal Volume, Fixed Volume, and Flow Rate. The C2H4 Offset, C2H4 Gain, Internal Volume values are set during calibration at Felix Instruments, and should not be changed unless instructed by a Felix Instruments technician. At the factory, the ethylene sensor is calibrated at 0 ppm and 20 ppm, then verification at 10 ppm.
The Fixed Volume and Flow Rate may be changed to accommodate various applications, but this may affect the accuracy of the device. Flow rate is adjustable from 40-120 mL/min. Fixed volume sample size is adjustable up to 18.5 mL.
The default Fixed Volume is 15 mL. The default Flow Rate is 67 mL/min.
Calibration
C2H4 Offset 0001672 C2H4 Gain 0000545 Internal Volume 016.2 Fixed Volume 15.0 Flow Rate 067
F-950 Instruction Manual rev. 10/17/16
1554 NE 3rd Ave, Camas, WA 98607, USA 27 Phone: (360) 833-8835
sales@felixinstruments.com www.felixinstruments.com
For additional information on gas calibration of the F-950 using the controller software, navigate to the FAQ on the F-950 support webpage:
https://felixinstruments.com/support/F-950/
and select How do I calibrate the instrument?
Or refer to Appendix I in this manual.
Setup Set Zero
Setup Set Zero sets a new baseline, or zero, for the C2H4 sensor only. This should be done daily for the C2H4 sensor. To set a new zero, first connect the external conditioning tube filled completely with potassium permanganate (KMnO4) to the inlet and use the “male to male adapter” to connect to the outlet of the F-950. Then use the right arrow to enter Set Zero.
Always allow the instrument adequate time to warm up (3 min) with the KMnO4 attached before setting the zero.
F-950 Instruction Manual rev. 10/17/16
28 1554 NE 3
rd
Ave, Camas, WA 98607, USA
Phone: (360) 833-8835
sales@felixinstruments.com
www.felixinstruments.com
Always store the KMnO4 air tight by connecting the ends of the tubing. Replace the cotton ball filters at either end of the external conditioning tube each time the KMnO4 is replaced (color indicator turns from purple to brown when it expires).
The instrument will run an automated process that sets a new baseline for future measurements. A reminder to use connect the potassium permanganate appears.
After the Set Zero routine is finished, it will exit to the Setup menu. Next, verify the set zero calibration by measuring the KMnO4 (0 ppm). Run the set zero calibration again if verification is not successful.
To set zero for CO2 and O2 (every 6 months), you will need to use Controller Software. This process is outlined in the How do I calibrate the instrument? Support section linked on the previous page.
Zero C2H4 Sensor
Use KMnO4 filter
C2H4 offset…..1349.1
F-950 Instruction Manual rev. 10/17/16
1554 NE 3rd Ave, Camas, WA 98607, USA 29 Phone: (360) 833-8835
sales@felixinstruments.com www.felixinstruments.com
Setup Date and Time
Use the Right and Left arrows to move between Month/Day/Year and Hour/Minute/Second and use the Up and Down arrows to change the values. To exit, use the left arrow to back out of the screen and return to the Setup menu.
F-950 Instruction Manual rev. 10/17/16
30 1554 NE 3
rd
Ave, Camas, WA 98607, USA
Phone: (360) 833-8835
sales@felixinstruments.com
www.felixinstruments.com
Setup RH Conversion
The Setup RH Conversion menu is used to correct for the temperature sensor being inside F-950 housing. This causes the temperature sensor to reflect the temperature of the F-950 and not the ambient temperature. The temperature is usually several degrees above ambient. The Relative Humidity is calculated based on temperature.
Enter the correct ambient temperature in degrees Celsius, measured from an external temperature sensor, to have the F­950 use for calculating the relative humidity (RH %). To use the
entered temperature for RH, “use sample T” should be set to “Yes.”
The current (internal RH) and corrected (sample RH) relative humidity are displayed at the bottom of this screen.
RH Conversion Enter Sample T(C) 20.0 Use Sample T Yes (for measurements)
Internal RH (%) 43.6 Sample RH (%) 21.3
F-950 Instruction Manual rev. 10/17/16
1554 NE 3rd Ave, Camas, WA 98607, USA 31 Phone: (360) 833-8835
sales@felixinstruments.com www.felixinstruments.com
Setup GPS
The GPS sensor inside of the F-950 can be used to record latitude and longitude to within 10 meters. The instrument should be operated outside, without overhead obstruction for best GPS performance. It may be difficult to acquire a fix on GPS satellites indoors. The Setup>GPS menu turns on or off the GPS sensor and show the current GPS data. Data is also saved to the SDcard .csv file.
Setup Language
The F-950 Three Gas Analyzer now has options for Spanish and Portugues languages for firmware 1.5.7.5 and above, to update your firmware refer to the Firmware Update portion of this manual.
Setup Factory Settings
The Setup > Factory Settings menu is used to back up or restore the factory defaults for the calibration parameters for all of the sensors (C2H4, CO2 and O2) on the device. The calibration parameters can also be modified using this menu. It is recommended to always back up the current parameters before making any modifications.
Setup GPS
Enable GPS Yes/No
Acquiring GPS data…..
Longitude 122.558
Latitude 45.59
F-950 Instruction Manual rev. 10/17/16
32 1554 NE 3
rd
Ave, Camas, WA 98607, USA
Phone: (360) 833-8835
sales@felixinstruments.com
www.felixinstruments.com
To modify calibration parameters:
1. Back up the current parameters (Setup>factory
settings>back up)
2. Make the modification in the .cfg fil
3. Import the .cfg file (Setup>factory settings>restore) to
apply the changes
File
The File Menu is accessed by highlighting “File” on the Main
Menu and pressing the right arrow key. Here, the user can manipulate files on the F-950. The F-950 remembers the last open file and will being saving data into it when it is powered back on.
In the main file menu, four options are available: Select, New, Delete and Review. All files created by the F-950 are .csv (comma separated value) files.
To view data on a computer, simply insert the SDcard into the computer’s SDcard reader. The computer should automatically detect the SDcard as a new storage device and mount the drive so that measurement data will be accessible by any computer application. The mini-USB port can also be used to establish a USB connection with a computer to transfer data from the F-
950.
F-950 Instruction Manual rev. 10/17/16
1554 NE 3rd Ave, Camas, WA 98607, USA 33 Phone: (360) 833-8835
sales@felixinstruments.com www.felixinstruments.com
File Select
File Select displays a list of .csv files that exist on the F-950 SD card. Use the Up and Down arrows to move between files, and the right arrow to select a file to which new data will be saved. If the unit is powered on and no file is selected, the data will be default saved to the file “data.csv”. In the file, each data point is labeled with time and date for easy sorting.
F-950 Instruction Manual rev. 10/17/16
34 1554 NE 3
rd
Ave, Camas, WA 98607, USA
Phone: (360) 833-8835
sales@felixinstruments.com
www.felixinstruments.com
File New
Pressing the Right arrow when File New is selected will create a new file according to the naming scheme programmed on the instrument, XX_XX_XX_X or Year_Month_Date_Ordinal. For example, the first file created on September 8, 2014 will read 14_09_08_0 and subsequent files will increase the last placeholder numerically. After pressing the right arrow, go to File Select to see the list of files on the SDcard. There will be a new file in the list with the current date.
File Delete
File Delete displays a list of files that exist on the F-950 SDcard. Use the Up and Down arrows to scroll between files and use the right arrow to display the option to delete the selected file. A message will appear that reads “Delete File?” Press the Left arrow for No, leaving the file intact. Press the Right arrow for Yes, deleting the file.
F-950 Instruction Manual rev. 10/17/16
1554 NE 3rd Ave, Camas, WA 98607, USA 35 Phone: (360) 833-8835
sales@felixinstruments.com www.felixinstruments.com
File Review
File Review displays a list of files that are on the SD card and allows you to view the data in the files. Use the Up and Down arrows to scroll between files and the right arrow to enter the selected file. The data.csv file is the default file, which data will be saved into if no other files are created.
After selecting a file name, the measurement mode with time of the measurement and ethylene concentration will appear for Fixed Volume or Trigger mode readings. Use the Up and Down arrows to highlight a measurement and the Right arrow to enter the measurement and see the more detailed data, including CO2, O2, Temperature, Relative Humidity and Flow Rate.
F-950 Instruction Manual rev. 10/17/16
36 1554 NE 3
rd
Ave, Camas, WA 98607, USA
Phone: (360) 833-8835
sales@felixinstruments.com
www.felixinstruments.com
Data Files on the Computer
Open the data files saved on the SDcard on the computer using Microsoft Excel or Notepad. Data files are saved as .csv (commas separated value). The following figure is an example data spreadsheet. Data values included are the date and time of the measurement, the measurement mode, the ethylene level in parts per million (ppm), the CO2 concentration in percent, the O2 concentration in percent, the temperature of the gas stream in degrees Celsius, the relative humidity (RH) of the gas stream in percent, and the flow rate of the gas stream in milliliters (mL) per minute.
REMEMBER: Always save the data files to the computer before
making changes or starting analysis.
F-950 Instruction Manual rev. 10/17/16
1554 NE 3rd Ave, Camas, WA 98607, USA 37 Phone: (360) 833-8835
sales@felixinstruments.com www.felixinstruments.com
Example spreadsheet data of an F-950 measurement.
F-950 Instruction Manual rev. 10/17/16
38 1554 NE 3
rd
Ave, Camas, WA 98607, USA
Phone: (360) 833-8835
sales@felixinstruments.com
www.felixinstruments.com
Firmware Update
For the latest version of firmware, update software, and instructions, visit the F-950 Support webpage (http://www.felixinstruments.com/support/f-950-support).
To update the firmware, you will need:
1. F-950_Package_Setup32.exe (or F-
950_Package_Setup64.exe, under Windows 64 bit)
2. Firmware code file: FW_950_v.X.X.X.X.dfu
3. Mini-USB to USB cable to connect to F-950 to PC.
Installing the F-950 Driver
1. Launch F-950_Package_Setup.exe.
2. Set up device by powering on the F-950 and connecting
it to computer via USB port. Go to Setup>Mode>USB menu and select Virtual COM.
3. Install will prompt for DfuSe tool (required for firmware
upgrade).
F-950 Instruction Manual rev. 10/17/16
1554 NE 3rd Ave, Camas, WA 98607, USA 39 Phone: (360) 833-8835
sales@felixinstruments.com www.felixinstruments.com
4. Click Next, then enter your information into the window
below. Click Next in the next several steps.
5. If Windows prompts the following message, select
“Install this driver software anyway”:
F-950 Instruction Manual rev. 10/17/16
40 1554 NE 3
rd
Ave, Camas, WA 98607, USA
Phone: (360) 833-8835
sales@felixinstruments.com
www.felixinstruments.com
6. Device drivers and Software package installation is
completed. Under Windows’s Start menu, you should
be able to find shortcuts for Device’s Controller and
Firmware Upgrade Program.
7. F-950 Controller software requires Java runtime
environment. Installer will check for appropriate version on your computer. Please download and install Java at
http://www.java.com/ if requested.
Updating the Firmware
1. Download latest firmware (.dfu file) from Felix
Instruments website:
https://felixinstruments.com/support/F-950/software/
2. Open the F-950 Firmware Upgrade software on your pc.
3. With the unit powered off, connect the F-950 to the
computer using the mini-USB to USB cable.
4. Hold the Down Arrow button down on the F-950, then
press the Power button, then release both buttons. This puts the F-950 in DFU mode.
5. On the DFU software, from the drop down menu under
“Available DFU Devices”, select “STM Device in DFU Mode”. Note: when connecting device to your
F-950 Instruction Manual rev. 10/17/16
1554 NE 3rd Ave, Camas, WA 98607, USA 41 Phone: (360) 833-8835
sales@felixinstruments.com www.felixinstruments.com
computer for the first time, it may take a minute for the device to show up in the drop down menu.
6. At the bottom of the transfer software screen, click
“Choose…”.
7. A dialogue box labeled open will appear. Navigate to
the desktop of your computer and select the new .dfu firmware file to download.
F-950 Instruction Manual rev. 10/17/16
42 1554 NE 3
rd
Ave, Camas, WA 98607, USA
Phone: (360) 833-8835
sales@felixinstruments.com
www.felixinstruments.com
8. Click “Upgrade”. A dialogue box will appear- click “Yes”
to proceed.
F-950 Instruction Manual rev. 10/17/16
1554 NE 3rd Ave, Camas, WA 98607, USA 43 Phone: (360) 833-8835
sales@felixinstruments.com www.felixinstruments.com
9. The device is now being updated with the newest
firmware. The green bar at the bottom of the screen tracks the progress of the upgrade. When the green bar reads “Upgrade successful” the process is complete.
10. Near the top of the DFU program, click “Leave DFU
mode” to return the device to its normal mode of
operation.
F-950 Instruction Manual rev. 10/17/16
44 1554 NE 3
rd
Ave, Camas, WA 98607, USA
Phone: (360) 833-8835
sales@felixinstruments.com
www.felixinstruments.com
11. Power on the F-950. In the lower right hand corner of
the device screen, the new version of firmware will appear.
F-950 Instruction Manual rev. 10/17/16
1554 NE 3rd Ave, Camas, WA 98607, USA 45 Phone: (360) 833-8835
sales@felixinstruments.com www.felixinstruments.com
F-950 Menu System Diagram
Below is a map of the F-950 ethylene analyzer menu system. Press the right arrow to enter a menu and the left arrow to exit.
F-950 Instruction Manual rev. 10/17/16
46 1554 NE 3
rd
Ave, Camas, WA 98607, USA
Phone: (360) 833-8835
sales@felixinstruments.com
www.felixinstruments.com
Appendix I: F-950 Controller Software and Sensor Calibration
The F-950 Controller software offers real-time monitoring, controlling as well as calibrating multiple F-950/F-900 devices from standard USB connection.
Instructions:
1. Open the F-950 Controller software
2. Connect the F-950 to the computer using the mini-USB
to USB cable.
3. In F-950 device Setup>Mode>USB menu, select
Controller mode.
F-950 Instruction Manual rev. 10/17/16
1554 NE 3rd Ave, Camas, WA 98607, USA 47 Phone: (360) 833-8835
sales@felixinstruments.com www.felixinstruments.com
4. In Controller software, the device series should show up
in the device list at top-left corner. Click on the device series to select device. Note: when connecting device to
your computer for the first time, it may take a minute for the device to show up in the left panel.
5. Monitoring information from your device should display
in monitor tab.
6. For more information on using the controller software
to calibrate the F-950, navigate to the FAQ page on the F-950 support webpage:
https://felixinstruments.com/support/F-950/
and select How do I calibrate the instrument?
Ethylene (C2H4) PPM Sensor Calibration
Setting the Zeroes (Offset) with Potassium Permanganate (KPMnO4) *daily
Takes approximately 5 minutes
1. Place fresh potassium
permanganate (KMnO4) in the external conditioning tube
2. Connect the conditioning tube to
the intake and outake of the F-950 utilizing Viton tubing (see photo)
3. Power on the F-950 and when
‘Measure’ is selected, right arrow
key and allow to run for 3 minutes to stabilize
4. Navigate back to the main menu and select Setup > Set
Zero, press the right arrow key to initiate.
5. When the instrument has set zero, a beep sounds and
the word ‘OK’ appears.
F-950 Instruction Manual rev. 10/17/16
48 1554 NE 3
rd
Ave, Camas, WA 98607, USA
Phone: (360) 833-8835
sales@felixinstruments.com
www.felixinstruments.com
Ethylene (C2H
4
) Sensor Set Span *6 months
1. Connect the high resolution flowmeter, Ethylene
calibration gas, and 950 intake ports together using a T­junction of tubing. If a sample bag of ethylene is being used, make a connection from the 950 intake port to the bag using the huber needle.
2. On the 950, begin measure mode.
3. Scroll through the graphs on the 950 (use the up and
down arrow keys) so that it is displaying the C2H4 concentration graph.
4. Set the flowrate on the C
2H4
gas tank so that the flow
meter reads roughly 70 mL/min. If the 950 gives a “Flow Blocked” error, then increase the flow on the meter by
10 mL/min and begin measure mode on the 950 again. If the error persists, contact tech support. If a sample bag is being used, skip this step.
5. Let the 950 read the gas for three minutes.
6. In the controller software, in the C2H4 field, set the
concentration to the value of the calibration gas being used. Then hit the “Set Span” button.
7. The concentration on the 950 should now read the
same as the calibration gas within 5% ± .5 ppm.
Ethylene (C2H
4
) Sensor Verification *After Setting the Span
Perform this process immediately after calibrating (set span) the ethylene sensor on the 950
1. Connect the high resolution flowmeter, Ethylene
verification gas, and 950 intake ports together using a T­junction of tubing. If a sample bag of ethylene is being
F-950 Instruction Manual rev. 10/17/16
1554 NE 3rd Ave, Camas, WA 98607, USA 49 Phone: (360) 833-8835
sales@felixinstruments.com www.felixinstruments.com
used, make a connection from the 950 intake port to the bag using the huber needle.
2. On the 950, begin measure mode.
3. Scroll through the graphs on the 950 (use the up and
down arrow keys) so that it is displaying the C2H4 concentration graph.
4. Set the flowrate on the C
2H4
gas tank so that the flow
meter reads roughly 90 mL/min. If the 950 gives a “Flow Blocked” error, then increase the flow on the meter by
10 mL/min and begin measure mode on the 950 again. If the error persists, contact tech support. If a sample bag is being used, skip this step.
5. Let the 950 read the gas for three minutes.
6. The 950 should read the same value as the verification
gas within 5% ± .5 ppm. If it does not, repeat the
“Ethylene Sensor Calibration” section above.
Carbon Dioxide (CO
2
) PCT Sensor Calibration
Setting the Zeroes (Offset) with Nitrogen Gas (N2) *6 months
Perform the set zero before the set span process. *User also has the ability to zero the sensor utilizing Soda Lime on closed loop and connecting to the controller software the same way.
1. Connect the high resolution flow meter, N
2
gas, and 950 intake ports together using a T-junction of tubing. If a sample bag of N2 is being used, make a connection from the 950 intake port to the bag using the huber needle.
F-950 Instruction Manual rev. 10/17/16
50 1554 NE 3
rd
Ave, Camas, WA 98607, USA
Phone: (360) 833-8835
sales@felixinstruments.com
www.felixinstruments.com
2. On the 950, begin continuous measure mode.
3. Set the flowrate on the N
2
gas tank so that the flow
meter reads roughly 90 mL/min. If the 950 gives a “Flow Blocked” error, then increase the flow on the meter by
10 mL/min and begin measure mode on the 950 again. If the error persists, contact tech support. If a sample bag of N2 is being used, skip this step.
4. Let the 950 read the gas for five minutes
5. In the controller software, select CO2 in the “Zero
sensors” field, then click the “Set Zero” button
6. The 950 should now read 0 for the CO
2
sensor
Carbon Dioxide (CO2) PCT Sensor Set Span *6 months
Perform the set zero before the set span process.
1. Connect the high resolution flow meter, CO
2
calibration gas, and 950 intake ports together using a T-junction of tubing. If a sample bag of CO2 is being used, make a connection from the 950 intake port to the bag using the huber needle.
2. On the 950 begin measure mode.
3. Scroll through the graphs on the 950 (use the up and
down arrow keys) so that it is displaying the CO2 concentration graph.
4. Set the flowrate on the CO
2
gas tank so that the flow
meter reads roughly 90 mL/min. If the 950 gives a “Flow Blocked” error, then increase the flow on the meter by
10 mL/min and begin measure mode on the 950 again.
F-950 Instruction Manual rev. 10/17/16
1554 NE 3rd Ave, Camas, WA 98607, USA 51 Phone: (360) 833-8835
sales@felixinstruments.com www.felixinstruments.com
If the error persists, contact tech support. If a sample bag is being used, skip this step.
5. Let the 950 read the gas for three minutes.
6. In the controller software, in the CO2 field, set the
concentration to the value of the calibration gas being
used. Then hit the “Set Span” button.
7. The concentration on the 950 should now read the
same as the calibration gas within 5%.
Carbon Dioxide (CO
2
) PCT Sensor Verification *After Set Span
Perform this process immediately after calibrating the carbon dioxide sensor on the 950
1. Connect the high resolution flow meter, CO
2
verification gas, and 950 intake ports together using a T-junction of tubing. If a sample bag of CO2 is being used, make a connection from the 950 intake port to the bag using the huber needle.
2. On the 950 begin measure mode.
3. Scroll through the graphs on the 950 (use the up and
down arrow keys) so that it is displaying the CO
2
concentration graph.
4. Set the flowrate on the CO
2
gas tank so that the flow
meter reads roughly 90 mL/min. If the 950 gives a “Flow Blocked” error, then increase the flow on the meter by
10 mL/min and begin measure mode on the 950 again. If the error persists, contact tech support. If a sample bag is being used, skip this step.
5. Let the 950 read the gas for three minutes.
F-950 Instruction Manual rev. 10/17/16
52 1554 NE 3
rd
Ave, Camas, WA 98607, USA
Phone: (360) 833-8835
sales@felixinstruments.com
www.felixinstruments.com
6. The 950 should read the same value as the verification
gas within 5%. If it does not, repeat the “Carbon Dioxide Sensor Calibration” section above.
Oxygen (O
2
) Sensor Calibration
Setting the Zeroes (Offset) with Nitrogen Gas (N2) *6 months
1. Connect the high resolution flow meter, N
2
gas, and 950 intake ports together using a T-junction of tubing. If a sample bag of N2 is being used, make a connection from the 950 intake port to the bag using the huber needle.
2. On the 950, begin continuous measure mode.
3. Set the flowrate on the N
2
gas tank so that the flow
meter reads roughly 90 mL/min. If the 950 gives a “Flow Blocked” error, then increase the flow on the meter by
10 mL/min and begin measure mode on the 950 again. If the error persists, contact tech support. If a sample bag of N2 is being used, skip this step.
4. Let the 950 read the gas for five minutes
5. In the controller software, select O2 in the “Zero
sensors” field, then click the “Set Zero” button
6. The 950 should now read 0 for the O
2
sensor
Oxygen (O2) Sensor Set Span *6 months
1. Connect the high resolution flow meter, O
2
calibration gas, and 950 intake ports together using a T-junction of
F-950 Instruction Manual rev. 10/17/16
1554 NE 3rd Ave, Camas, WA 98607, USA 53 Phone: (360) 833-8835
sales@felixinstruments.com www.felixinstruments.com
tubing. If a sample bag of O2 is being used, make a connection from the 950 intake port to the bag using the huber needle.
2. On the 950 begin measure mode.
3. Scroll through the graphs on the 950 (use the up and
down arrow keys) so that it is displaying the O2 concentration graph.
4. Set the flowrate on the O
2
gas tank so that the flow
meter reads roughly 90 mL/min. If the 950 gives a “Flow Blocked” error, then increase the flow on the meter by
10 mL/min and begin measure mode on the 950 again. If the error persists, contact tech support. If a sample bag is being used, skip this step.
5. Let the 950 read the gas for three minutes.
6. In the controller software, in the O2 field, set the
concentration to the value of the calibration gas being used. Then hit the “Set Span” button.
7. The concentration on the 950 should now read the
same as the calibration gas ± 5%.
Oxygen (O
2
) Sensor Verification *After Set Span
Perform this process immediately after calibrating the oxygen sensor on the 950
1. Connect the high resolution flow meter, O
2
verification gas, and 950 intake ports together using a T-junction of tubing. If a sample bag of O2 is being used, make a connection from the 950 intake port to the bag using the huber needle.
2. On the 950 begin measure mode.
F-950 Instruction Manual rev. 10/17/16
54 1554 NE 3
rd
Ave, Camas, WA 98607, USA
Phone: (360) 833-8835
sales@felixinstruments.com
www.felixinstruments.com
3. Scroll through the graphs on the 950 (use the up and
down arrow keys) so that it is displaying the O
2
concentration graph.
4. Set the flowrate on the O
2
gas tank so that the flow
meter reads roughly 90 mL/min. If the 950 gives a “Flow Blocked” error, then increase the flow on the meter by
10 mL/min and begin measure mode on the 950 again. If the error persists, contact tech support. If a sample bag is being used, skip this step.
5. Let the 950 read the gas for three minutes.
6. The 950 should read the same value as the verification
gas ± 5%. If it does not, repeat the “Oxygen Sensor Calibration” section above.
Continuous Mode Verification *After Set Span
The following procedure is to check whether the sensors were calibrated to read within spec of the actual gas value.
It is highly recommended that this step be performed after one full day has passed since the calibration was performed. If this step is not performed, accuracy of the calibration cannot
be verified.
Repeat the verification sections from above for each gas and record the results here after three minutes (at least) of measure time. If any of the values are out of specification, check your procedure and recalibrate the sensor again.
Gas
950 Results
± 5% of standard gas
Air (Air is ~20.9% O2)
□Yes □No
CO
2
Verification Gas
□Yes □No
O
2
Verification Gas
□Yes □No
F-950 Instruction Manual rev. 10/17/16
1554 NE 3rd Ave, Camas, WA 98607, USA 55 Phone: (360) 833-8835
sales@felixinstruments.com www.felixinstruments.com
Gas
950 Results
5% ± .5 PPM of
standard gas
C2H4 PPM Verification Gas
□Yes □No
For information on sourcing known gases required for calibration, please refer to the following website to inquire about your region:
https://www.airliquide.com/group/where-we-operate
F-950 Instruction Manual rev. 10/17/16
56 1554 NE 3
rd
Ave, Camas, WA 98607, USA
Phone: (360) 833-8835
sales@felixinstruments.com
www.felixinstruments.com
Appendix II: Sampling from a Jar
The following photos utilize a glass jar customized with the addition of a Vacutainer® septum. The F-950 can draw and analyze a headspace gas sample from a rigid jar by incorporating an input probe and output flow into the septum to compensate for removed airspace, thus preventing a vacuum from occurring.
Instructions:
1. With the F-950 in Fixed Volume Mode, allow the
instrument to purge and stabilize in ambient air before inserting the fixed volume probe into the sample jar septum.
2. Once stabilization is complete, insert the fixed volume
needle probe connected to the F-950 inlet into the jar septum.
3. To prevent a vacuum, insert another probe from the
outlet (bottom) of the F-950 into the septum. There are now two needle probes inserted into the septum.
4. Press the square button above the name of the
instrument to initiate the Fixed Volume measurement. The pump inside of the F-950 will activate and draw headspace gas into the unit.
5. Remove both needle probes during each purge to allow
the instrument to properly purge the internal tubing in the F-950.
F-950 Instruction Manual rev. 10/17/16
1554 NE 3rd Ave, Camas, WA 98607, USA 57 Phone: (360) 833-8835
sales@felixinstruments.com www.felixinstruments.com
1 3 5
*If you receive a Clogged probe? error, check that your needle is not bent and that the needle has penetrated through the septum and into the sampling space. For larger sampling jars, consider incorporating an internal fan to improve circulation within the sample space. Carbon dioxide gas can sink to the bottom portion of the jar, rendering a non­representative reading to occur. Placing the jar on a long side helps to mitigate this from affecting the sample draw and more evenly distributes the internal gas.
F-950 Instruction Manual rev. 10/17/16
58 1554 NE 3
rd
Ave, Camas, WA 98607, USA
Phone: (360) 833-8835
sales@felixinstruments.com
www.felixinstruments.com
F-950 Production Test Check Sheet
SERIAL NUMBER:
Firmware Version:
C2H4 CALIBRATION PARAMETERS
C2H4 Offset
C2H4 Gain
Internal Volume
Fixed Volume
Flow Rate
NOTES:
F-950 Instruction Manual rev. 10/17/16
1554 NE 3rd Ave, Camas, WA 98607, USA 59 Phone: (360) 833-8835
sales@felixinstruments.com www.felixinstruments.com
F-950 Instruction Manual rev. 10/17/16
60 1554 NE 3
rd
Ave, Camas, WA 98607, USA
Phone: (360) 833-8835
sales@felixinstruments.com
www.felixinstruments.com
FELIX INSTRUMENTS HARDWARE WARRANTY
Important: Please Read!
Seller’s Warranty and Liability: Seller warrants new equipment
of its own manufacturing against defective workmanship and materials for a period of one year, of a single shift operation, from date of receipt of equipment - the results of ordinary
wear and tear, neglect, misuse, accident and excessive deterioration due to corrosion from any cause is not to be considered a defect. Any defect must be called to the attention
of Felix Instruments, Camas, Washington, USA, in writing, within 90 days after receipt of the unit.
Seller’s liability for defective parts is limited to the repair or
replacement of any part of the instrument without charge, if Felix Instruments’ examination discloses that part to have been defective in material or workmanship, and in no event shall exceed the furnishing of replacement parts F.O.B. the factory where originally manufactured. No equipment may be repaired or altered by anyone not authorized by Felix Instruments.
Material and equipment covered hereby, which is not manufactured by Seller, is to be covered only by the warranty of its manufacturer. Seller shall not be liable to the Buyer for loss, damage, or injury to persons (including death), or to property or things, whatsoever, including, but without limitation, products processed by the use of the equipment; or for damages of any kind or nature (including, but without limitation, loss of anticipated profits), occasioned by or arising out of installation, operation, use, misuse, nonuse, repair, or replacement of said
F-950 Instruction Manual rev. 10/17/16
1554 NE 3rd Ave, Camas, WA 98607, USA 61 Phone: (360) 833-8835
sales@felixinstruments.com www.felixinstruments.com
material and equipment, or out of the use of any method or process for which the same may be employed. The purchaser is to pack, ship, or deliver the instrument to Felix Instruments, in Camas, Washington, USA, within 30 days after Felix Instruments has received written notice of the defect at the customer’s expense. No other arrangements may be made unless otherwise approved in writing by Felix Instruments.
The use of this equipment constitutes Buyer’s acceptance of the
terms set forth in this warranty. There are no understandings, representations, or warranties of any kind, express, implied, statutory, or otherwise (including, but without limitation, the
implied warranties of merchantability and fitness for a particular purpose), not expressly set forth herein.
F-950 Instruction Manual rev. 10/17/16
62 1554 NE 3
rd
Ave, Camas, WA 98607, USA
Phone: (360) 833-8835
sales@felixinstruments.com
www.felixinstruments.com
F-950 Instruction Manual rev. 10/17/16
1554 NE 3rd Ave, Camas, WA 98607, USA 63 Phone: (360) 833-8835
sales@felixinstruments.com www.felixinstruments.com
Warranty Registration Card
1554 NE 3rd Ave, Camas, WA 98607, USA Phone: (360) 833-8835 Fax: (360) 833-1914 e-mail: sales@felixinstruments.com Web: www.felixinstruments.com
PRODUCT REGISTRATION CARD
Please complete and return this form to Felix Instruments within 30 days to validate your Warranty on Parts & Labor.
Registration Information:
Your Name:____________________________________ Title:__________________
Company/University:___________________________________________________
Address:_____________________________________________________________
City:____________________________ State:__________ Zip:__________________
Country:__________________________Email_______________________________
Phone:______________________________ Fax:____________________________
Felix Instruments Serial Number(s):__________________________________________________
Purchase Date:_____________________Purchase Price:_____________________
------------------------------------------------------------------------------------------------------------------
FOLD ON DOTTED LINE
Your opinions will help improve our service. Please answer the following questions.
1. What was the basis of your product selection?
Representative Recommendation Price Product Features Product Design Technical Specifications Brand Name Warranty Service Other _________________________________________
2. What other competing brands did you consider? __________________________
3. Where did you first learn of this product?
Advertisement in ________________________________ □ Representative Friend/Colleague Exhibit Other _________________________________________
4. Who selected this product?
I did Research Group University Department Purchasing Other _________________________________________
5. Comments/Suggestions:
______________________________________________________________________
______________________________________________________________________