Fazzini F-18, F-20, F-36, F-40 User manual

Suction Pump
F-18/F-36/F-20/F-40
Service Manual
NOTE
This manual is related to the following Equipment:
Suction Pump
F-18 / F-36 F-20 / F-40
Suction Pump
F-18/F-36/F-20/F-40
Service Manual
All medical device equipment and accessories should be in good working order prior to use. It is the responsibility of each facility’s engineering department to test equipments and “routine basis” to
determine if medical devices are functioning within specifications.
No device or accessory should be used if a malfunction is suspected.
The installation, operation, search for breakdowns and the repair of equipment must only be performed by trained personnel
supplied.
1
( )
who have sufficient knowledge about the equipment, spare parts, and the information
BEFORE STARTING TO INSPECT THE DEVICE, MAKE SURE IT IS NOT CONNECTED TO THE
ELECTRIC POWER SUPPLY NETWORK.
DO NOT USE THE APPLIANCE IN THE PRESENCE OF INFLAMMABLE AND/OR EXPLOSIVE
LIQUIDS, GASES OR MIXTURES.
DO NOT LEAVE THE APPLIANCE EXPOSED TO ATMOSPHERIC AGENTS OR HEAT SOURCES.
USING THE APPLIANCE UNDER DIFFERENT CONDITIONS THAN THOSE SPECIFIED IN THIS
HANDBOOK MAY PREJUDICE ITS SAFETY AND ITS TECHNICAL PARAMETERS
CAUTIONS
1.
An educated (qualified) person has sufficient knowledge or experience to allow him to avoid the dangers presented by electricity, whereas an informed (trained)
person is sufficiently informed or works supervised by qualified persons, so as to allow him to avoid the dangers presented by electricity (e.g. operators or maintenance staff).
Suction Pump
F-18/F-36/F-20/F-40
Service Manual
BEFORE USING THE APPLIANCE READ THE HANDBOOK
THE APPLIANCE MUST BE USED BY QUALIFIED PERSONNEL
PLEASE CONTACT THE FAZZINI S.R.L. TECHNICAL DIVISION FOR ANY PROBLEM SHOULD ARISE
CAREFULLY FOLLOW THE GUIDELINES WHEN USING THE APPLIANC E:
DO NOT HANDLE THE APPLIANCE WITH WET OR MOIST HANDS DO NOT HANDLE THE PLUG WITH WET OR MOIST HANDS DO NOT USE THE APPLIANCE UNDER THE SHOWER OR IN THE BATHTUB. DO NOT DIP THE APPLIANCE UNDER WATER DO NOT USE THE APPLIANCE IN THE PRESENCE OF INFLAMMABLE AND/OR EXPLOSIVE LIQUIDS, GASES OR MIXTURES DO NOT LEAVE THE APPLIANCE EXPOSED TO ATMOSPHERIC AGENTS OR HEAT SOURCES DO NOT LET CHILDREN OR INCAPABLE PERSONS USE THE APPLIANCE WITHOUT THE NECESSARY SUPERVISION. BEFORE STARTING TO CLEAN THE DEVICE, MAKE SURE IT IS NOT CONNECTED TO THE ELECTRIC POWER SUPPLY NETWORK DO NOT USE THE APPLIANCE FOR AIMS OTHER THAN THOSE FORESEEN BY THE MANUFACTURER. USING THE APPLIANCE UNDER DIFFERENT CONDITIO NS THAN THOSE SPECIFIED IN THIS HANDBOOK MAY PREJUDICE ITS SAFETY AND ITS TECHNICAL PARAMETERS
MODEL
LENGTH
WIDTH HEIGHT WEIGHT
POWER SUPPLY
MAX ASPIRATION
FLOW
-WITH FILTER
-DIRECT
MEASUREMENT ON
ASPIRATION VENT
FUSE
DO NOT OPEN THIS DEVICE FOR ANY REASON
F-18.XX
260 MM. 180 MM. 220 MM.
3,4 KG
230V- 50/60Hz
- 0,75 BAR
11 L/MIN.
18 LT/MIN
1 X 1,6A 250V

WARNINGS

WARNINGS

WARNINGS

CAUTIONS

TECHNICAL DATA

F-36.XX
260 MM. 180 MM. 220 MM.
3,9 KG
230V- 50/60Hz
- 0,75 BAR
16 L/MIN.
36 LT/MIN
1 X 1,6A 250V
F-20.XX
350 MM. 210 MM. 250 MM.
4.0 KG
230V- 50/60Hz
- 0,75 BAR
11 L/MIN.
18 LT/MIN
1 X 1,6A 250V
F-40.XX
350 MM. 210 MM. 250 MM.
4.5 KG
230V- 50/60Hz
- 0,75 BAR
16 L/MIN.
36 LT/MIN
1 X 1,6A 250V
THE APPLIANCE IS A PORTABLE ASPIRATOR FOR USE IN MINOR SURGERY AND AIMED AT FREEING THE AIRWAYS FROM ORGANIC SECRETIONS.
OPERATIONAL TEMPERATURE: BETWEEN 0° AND 50° OPERATIONAL HUMIDITY: BETWEEN 20 AND 90% RU BEYOND 2500 METERS OF HEIGHT, THE ASPIRATION FLOW MAY DECREASE.
1.TIGHTLY SCREW THE LID ON THE JAR TO GUARANTEE A PERFECT SEAL
2.CONNECT THE SHORT TUBE TO FILTER 1 ON THE ‘’ASPIRATION’’ 2 VENT
3.CONNECT THE OTHER END OF THE TUBE TO THE ‘’VACUUM’’ 3 VENT WITH THE FLOAT ON THE JAR’S LID
4.CONNECT THE LONG TUBE 4 TO THE OTHER ‘’PATIENT’’ 5 VENT ON THE JAR’S LID (VENT WITHO UT FLOAT)
5.CONNECT THE OTHER END OF THE LONG TUBE 4 THE PROBE CONNECTO R 6 AND EVENTUALLY THE SUPPLIED PROBE
6. PLUG IN THE POWER CORD TO THE MAIN BUT ONLY AFTER CHECKING THE VOLTAGE IS THE ONE INDICATED BY THE LABEL IN THE BACK OF THE EQUIPMENT.
7.PRESS THE ON/OFF SWITCH 7
8.ASPIRATION INTENSITY CAN BE VARIED WITH THE RELEVANT CONTROL ON THE FRONT PANEL OF THE APPLIANCE 8 AND CAN BE CHECKED ON THE APPLIANCE FRONT PA NEL 9 VACUOMETER.
9.IT IS RECOMMENDED TO CARRY OUT A FIRST ASPIRATION CYCLE WITH AT LEAST 500 ML OF WATER ONLY.
10.THE ASPIRATED FLUID WILL FLOW IN THE RELEVANT JAR. SHOULD THE AMOUNT OF FLUID EXCEED THE JAR’S CAPACITY, THE OVERFLOW VALVE WILL STOP ASPIRATION.
11.IMMEDIATELY SWITCH OFF THE APPLIANCE BY SETTING THE ON/OFF 7 SWITCH ON THE ‘’ OFF’’ POSITION. THE MANUFACTURER, ASSEMBLER, INSTALLER OR IMPORTER ARE HELD LIABLE FOR THE SAFETY, PERFORMANCE AND
RELIABILITY OF THE APPLIANCE ONLY IF THE ELECTRIC SYSTEM TO WHICH THE APPLIANCE IS CONNECTED COMPLIES WITH L.D.46/90
ACCESSORIES INCLUDED CODE N°
05.0001
05.00552
05.0010
05.0015
SHOULD ACCESSORIES BE USED OTHER THAN THOSE SUPPLIED BY FAZZINI S.R.L., YOU MUST MAKE SURE THAT THEY HAVE THE SAME CHARACTERISTICS AS THE ORIGINAL ONES.
NOTE: THE SUPPLIED JARS ARE NOT MEANT TO BE USED AS MEASURERS, AND P ROVIDE ONLY AN INDICATION OF THE CONTAINED QUANTITY.

OPERATIONAL STEPS AT THE END OF THE PROCEDURE

NOTE: ALWAYS USE INDIVIDUAL PROTECTION DEVICES (EX. LATTICE GLOVES) EVERY TIME YOU MAY GET IN CONTACT WITH POTENTIALLY CONTAMINATED PARTS. DISCONNECT DISPOSABLE PARTS AND DISPOSE OF THEM. CHECK THE GENERAL INTEGRITY OF THE ASPIRATOR CARRY OUT THE FOLLOWING CLEANING OPERATIONS ASPIRATION TUBES SHOULD BE STERILIZED AFTER EACH USE.
THE FILTER ON THE APPLIANCE IS DISPOSABLE; THEREFORE IT IS NECESSARY TO REPLACE I T AFTER EVERY USE. DO NOT USE THE ASPIRATOR WITHOUT A FILTER
THE FLUID COLLECTION JAR CAN BE AUTOCLAVED AT A TEMPERATURE NOT EXCEEDING 121ºC, OTHERWISE, THE JAR CAN BE DISINFECTED WITH DENATURED ALCOHOL OR A HYPO CHLORITE BASED SOLUTION THAT CAN BE EASILY FOUND IN PHARMACIES MAXIMUM REUSE LIMIT: 30 CYCLES
Suction Pump
F-18/F-36/F-20/F-40
Service Manual

OPERATING

DESCRIPTION JAR 1000 ML SET OF TUBES ASPIRATION PROBE CH 14 (DISPOSABLE) ANTIBACTERIAL/HYDROPHOBIC FILTER (DISPOSABLE)

JAR CLEANING

Loading...