Fazzini F-30 Service manual

Page 1
Suction Unit
Service Manual
NOTE
This manual is related to the following Equipment:
Suction Pump
F-30.xx
F-30.xx
Doc : Service Manual - F-30.doc Rev. 2 / 12 APR 2006 Page 1 of 10
Page 2
Suction Unit
F-30.xx
Service Manual
All equipment and accessories should be in good working order prior to use.
It is the responsibility of each facility’s engineering department to test equipments and “routine basis” to determine if the devices are functioning within specifications.
No device or accessory should be used if a malfunction is suspected.
The installation, operation, search for breakdowns and the repair of equipment must onl y be performed by traine d personnel
BEFORE STARTING TO INSPECT THE DEVICE, MAKE SURE IT IS NOT CONNECTED TO THE ELECTRIC POWER SUPPLY NETWORK.
DO NOT USE THE APPLIANCE IN THE PRESENCE OF INFLAMMABLE AND/OR EXPLOSIVE LIQUIDS, GASES OR MIXTURES.
DO NOT LEAVE THE APPLIANCE EXPOSED TO ATMOSPHERIC AGENTS OR HEAT SOURCES.
USING THE APPLIANCE UNDER DIFFERENT CONDITIONS THAN THOSE SPECIFIED IN THIS HANDBOOK MAY PREJUDICE ITS SAFETY AND ITS TECHNICAL PARAMETERS
DELIVERY AND UNCRATING OF UNIT
Inspect equipment and shipping crate immediately upon receipt. If any damage is apparent, you should both report it to the trucking delivery person and contact the transportation company immediately.
Make notes of any damage on the Bill Of Lading. Retain all shipping materials for ins pection. Any claims for damage must start at the receiving point. Do not return goods without written authorization.
When submitting a claim for shipping damage, request that the carrier inspect the shipping container and equipment.
Check packing slip carefully and make sure all materials have been received as indicated on the packing ticket.
(1)
who have sufficient knowledge about the equipment, spare parts, and the information supplied.
CAUTIONS
1.
An educated (qualified) person has sufficient knowledge or experience to allow him to avoid the dangers presented by electricity, whereas an informed (trained)
person is sufficiently informed or works supervised by qualified persons, so as to allow him to avoid the dangers presented by electricity (e.g. operators or maintenance staff).
Doc : Service Manual - F-30.doc Rev. 2 / 12 APR 2006 Page 2 of 10
Page 3
Suction Unit
TECHNICAL SPECIFICATIONS
LENGTH mm 450
WIDTH mm 150
HEIGHT mm 280
WEIGHT kg 4,00
POWER SUPPLY 230V – 50/60 Hz
MAX VACUUM bar - 0,75
MAX FLOW – WITH FILTER l/mins 10
MAX FLOW –DIRECT MEASUREMENT ON ASPIRATION VENT l/mins 40
FUSES A 2x1,25
ACCESSORIES INCLUDED
CODE N° DESCRIPTION
05.0001 OR 05.0002 JAR 1000 ML OR 2000 ML
05.00552 SET OF TUBES
05.0010 ASPIRATION PROBE CH 14 (DISPOSABLE)
05.0016 ANTIBACTERIAL/HYDROPHOBIC FILTER (DISPOSABLE)
Doc : Service Manual - F-30.doc Rev. 2 / 12 APR 2006 Page 3 of 10
F-30.xx
Service Manual
Page 4
Suction Unit
OPERATING The appliance is a portable SUCTION UNIT FOR EMERGENCY USE, MINOR SURGERY AND aimed at freeing
the airways from organic secretions.
OPERATIONAL TEMPERATURE: BETWEEN 0° AND 50° RECOMMENDED RECHARGE TEMPERATURE: BETWEEN 15° AND 30° OPERATIONAL HUMIDITY: BETWEEN 20 AND 90% RU
BEYOND 2500 METERS OF HEIGHT, THE ASPIRATION FLOW MAY DECREASE. THE FOLLOWING STEPS ARE POSSIBLE ONLY IF THE APPLIANCE IS COMPLETELY RECHARGED
1.TIGHTLY SCREW THE LID ON THE JAR TO GUARANTEE A PERFECT SEAL
2.CONNECT THE SHORT TUBE TO FILTER 1 ON THE ‘’ASPIRATION’’ 2 VENT
3.CONNECT THE OTHER END OF THE TUBE TO THE ‘’VACUUM’’ 3 VENT WITH THE FLOAT ON THE JAR’S
LID
4.CONNECT THE LONG TUBE 4 TO THE OTHER ‘’PATIENT’’ 5 VENT ON THE JAR’S LID (VENT WITHOUT FLOAT)
5.CONNECT THE OTHER END OF THE LONG TUBE 4 TO THE PROBE CONNECTOR 6 AND EVENTUALLY THE SUPPLIED PROBE
7.PRESS THE ON/OFF SWITCH 7
8.ASPIRATION INTENSITY CAN BE VARIED WITH THE RELEVANT CONTROL ON THE FRONT PANEL OF THE APPLIANCE 8 AND CAN BE CHECKED ON THE APPLIANCE FRONT PANEL 9 VACUOMETER.
9.IT IS RECOMMENDED TO CARRY OUT A FIRST ASPIRATION CYCLE WITH AT LEAST 500 ML OF WATER
ONLY.
10.THE ASPIRATED FLUID WILL FLOW IN THE RELEVANT JAR. SHOULD THE AMOUNT OF FLUID EXCEED THE JAR’S CAPACITY, THE OVERFLOW VALVE WILL STOP ASPIRATION.
11.IMMEDIATELY SWITCH OFF THE APPLIANCE BY SETTING THE ON/OFF 7 SWITCH ON THE ‘’ OFF’’ POSITION.
Doc : Service Manual - F-30.doc Rev. 2 / 12 APR 2006 Page 4 of 10
F-30.xx
Service Manual
Page 5
Suction Unit
F-30.xx
Service Manual
PERIODIC MAINTENANCE CHECK
SUCTION UNIT WAS DESIGNED TO KEEP MAINTENANCE TO A MINIMUM. NEVERTHELESS IT IS NECESSARY TO CARRY OUT A DAILY CHECK BEFORE USING THE APPLIANCE.
BEFORE CARRYING OUT ANY CHECK DUE TO MALFUNCTIONING OR ANOMALIES, CALL THE FAZZINI S.R.L. TECHNICAL SERVICE OR PERSONNEL QUALIFIED BY FAZZINI S.R.L.
BATTERY CAN BE CHANGED ONLY BY TRAINED PEOPLE FAZZINI S.R.L. DOES NOT OFFER ANY WARRANTY ON APPLIANCE THAT UPON INSPECTION BY THE
TECHNICAL SERVICE SHOWS SIGNS OF TAMPERING. THE OPERATOR MUST EXCLUSIVELY REPLACE THE SUPPLIED ACCESSORIES IF NECESSARY IN LINE
WITH THE APPLIANCE’S USE INTENSITY ASPIRATION TUBES SHOULD BE STERILIZED AFTER EACH USE. IN CASE OF REPAIR, THE USER
SHOULD CALL FAZZINI S.R.L. TRANSPORTATION COSTS MUST ALWAYS BE AGREED WITH FAZZINI S.R.L. BEFORE SHIPPING FAILED PIECES.
SHOULD ANY FLUID LEAK IN THE APPLIANCE, CALL THE TECHNICAL SERVICE OF FAZZINI S.R.L. FOR MAINTENANCE. DO NOT OPEN THE UNIT.
STORAGE
THE SUCTION UNIT MUST BE STORED IN A DRY LOCATION, PROTECTED FROM ATMOSPHERIC AGENTS STORAGE TEMPERATURE: BETWEEN -10° AND + 55° RELATIVE HUMIDITY BETWEEN 20 AND 80%
Doc : Service Manual - F-30.doc Rev. 2 / 12 APR 2006 Page 5 of 10
Page 6
Suction Unit
F-30.xx
Service Manual
Legenda F-30.xx
Doc : Service Manual - F-30.doc Rev. 2 / 12 APR 2006 Page 6 of 10
1 Filter 2 Vacuum drive 3 Jar drive to the filter 4 Disposable tube to the patient 5 Jar drive to the patient 6 Suction catheter CH 14 7 Main switch 8 Vacuum regulator 9 Vacuum gauge
Page 7
Suction Unit
F-30.xx
Service Manual
General assembly – F-31.xx
Doc : Service Manual - F-30.doc Rev. 2 / 12 APR 2006 Page 7 of 10
Page 8
Suction Unit
F-30.xx
Service Manual
Functional layout
Main
Doc : Service Manual - F-30.doc Rev. 2 / 12 APR 2006 Page 8 of 10
Page 9
Suction Unit
F-30.xx
Service Manual
Pneumatic circuit
2
1
3
4
5
6
7
1 Catheter to the patient 2 Suction regulator 3 Jar 4 Non rebeathing valve 5 Filter 6 Pump / Motor 7 Vent
Doc : Service Manual - F-30.doc Rev. 2 / 12 APR 2006 Page 9 of 10
Page 10
Suction Unit
F-30.xx
Service Manual
Troobleshooting
Situation Possible Cause Solution
a Disconnected or loose cord Check connections and reconnect unit
b Faulty wall outlet Use another wall outlet
Equipment will not
1
turn on
Situation Possible Cause Solution
c Fuses are damaged
d Main rocker switch is damaged
Internal wiring connection are not
e
completed
Vacuum regulator do not work
a
properly
Check the fuses inside SP.38.02
If fuses are damaged check the electrical insulation
on the electrical apparatus
In ON position, check the presence of the light ; if no light is detected replace with a spare switch
Open the equipment and verify internal electrical wiring (see functional layout)
Turn the manifold SP.21
2 Suction fault
Internal tubing are not properly
b
connected
c Internal tubing are choked Open the equipment and verify internal tubing bends
d Pump fault
Open the equipment and verify internal tubing connection
Open the equipment ,disconnect the tubing and check the pump
Doc : Service Manual - F-30.doc Rev. 2 / 12 APR 2006 Page 10 of 10