Cino L780, F680, L680, F790, F780 Programming Manual

...
Page 1
International Edition, Rev. D14
Page 2
FuzzyScan
Revision History
Rev. No.
Released Date
Description
Rev.B Beta
Apr. 16, 2009
First Release
Rev.B0
May 05, 2009
Modify Symbology Reading Control - Code 128/EAN-128 Setting and GS1 DataBar Setting Add Appendix - Symbology ID Table.
Rev.B1
June 22, 2009
Add Operation Control - Presentation Scanning Mode.(Rename to Presentation Control on
Rev.B4)
 Add Operation ControlTime Delay to Low Power Mode.
Rev.B2
Aug. 21, 2009
Add “Symbology Reading Control” - “Composite Code Setting”, “PDF/Micro PDF417 Setting”,
Codablock F Setting, Korea Post Code Setting”.
Add “Operation Control” – “SmartStand Power Off Timeout”. Add “Condensed DataWizard” - “PDF417/Micro PDF417”, “Codablock F”, “Korea Post Code” Modify “Appendix” – “ Symbology ID Table”
Rev.B3
Oct. 16, 2009
Add “Host Interface Selection” - “IBM PS/2, 25-30 series keyboard wedge interface”. Add “Operation Control” - “Presentation Scanning Auto-sense”. Add “Appendix” – “ Symbology ID Table” – “Code 128”
Rev.B4
Mar. 05, 2010
Modify “Symbology Reading Control” - “UPC-A & UPC-E Setting” Rename “Serial Interface Control” – “Time Out Control” to “Serial Response Time-out”. Rename “ Operation Control” – “Presentation Scanning Mode” to “Presentation Control” Rename “Operation Control” – “Auto Power Off Duration” to “Light Source On Time”. Rename “Operation Control” – “ Presentation Scanning Auto-sense” to “Presentation Auto-sense” Modify “Operation Control” - “Good Read Duration” Add “Operation Control” - “ Presentation Sensitivity” Add ““Appendix” – “Master Default” System Command
Rev. B5
Apr.12, 2010
 Modify Operation Control - Hands Free Time-out
Page 3
FuzzyScan
Revision History
Rev. No.
Released Date
Description
Rev. B6
May 27, 2010
 Modify Serial Interface Control Baud Rate
Rev. B7
Jun 25 , 2010
Modify Symbology Reading Control - UPC-A & UPC-E Setting Modify Operation Control - "Time Delay to Low Power Mode"
Rev. B8
Sep 29 , 2010
Modify Symbology Reading Control - UPC-A & UPC-E Setting and EAN Setting Modify Keyboard Interface ControlKeyboard Layout (Language) Setting Add Operation Control - Scan Rate Control and Good Read Indicator
Rev. B9
Nov 10 , 2010
 Modify Serial Interface Control Baud Rate
Rev. C1
Mar 04, 2011
 Modify Serial Interface Control Baud Rate
Rev. C2
Mar 22, 2011
Modify Symbology Reading Control - Code 39 Setting Modify Serial Interface ControlProtocol, ACK/NAK Setting
Rev. C3
Jul 22, 2011
Rename “ Symbology Reading Control” - “ UCC/EAN-128” to “GS1-128” Rename ““Serial Interface Control” – “ ACK/NAK Transmission Indication” to “ACK Indication” Modify “Serial Interface Control” – “ACK Indication” and “Serial response time -out”. Rename “Condensed DataWizard” - “ UCC/EAN-128” to “GS1-128”
Rev. C4
Nov 04, 2011
 Modify Symbology Reading Control - Codabar/ NW-7 Setting
Rev. C5
Jan 02, 2012
 Modify Wand/Laser Emulation Control Code 39/Code 128 Emulation
Rev. C6
Mar 05, 2012
Add Operation Control - "1D Barcode Forward-reading Indication, 1D Barcode Backward-reading
Indication, and 1D Barcode Direction Indication Transmission”.
Rev. C7
Jul 25, 2012
Add “Host Interface Selection” – “ USB HID Legacy” Modify “Symbology Reading Control” - “Code 128 Setting” Modify “ Wand/Laser Emulation Control” – “ Code 39/Code 128 Emulation” Modify “Operation Control” - "Operation Mode Setting”, “Presentation Control” Add “Operation Control” - " LED illumination Control” and “LED Illumination Delay”
Rev. D1
Dec. 18, 2012
Support 2D Functions.
Page 4
FuzzyScan
Revision History
Rev. No.
Released Date
Description
Rev. D2
Apr. 22. 2013
Modify Symbology Reading Control - Data Matrix Setting Modify Operation Control - Batch Reading rule example
Rev. D3
Aug. 20, 2013
Modify Symbology Reading Control - Readable Bar Code Setting Modify Appendix - Symbology ID Table
Rev. D4
Sep. 06, 2013
 Modify Condensed DataWizard - 1D Bar Code Symbology and 2D Bar Code Symbology
Rev. D5
Mar. 06, 2014
Modify Symbology Reading Control - GS1 DataBar Setting Add Appendix USB HID Legacy Mode Quick Set
Rev. D6
Feb. 06, 2015
 Modify Symbology Reading Control - MaxiCode Setting
Rev. D7
Aug. 17, 2015
 Add Host Interface Selection USB EFT Terminal Mode
Rev. D8
Dec. 25, 2015
Revise Revision History Add Getting Started - A670 series scanner
Rev. D9
May 31, 2016
 Add Operation ControlBuzzer Volume
Rev. D10
Aug. 04, 2016
 Add Symbology Reading Control - Small DM Code Reading
Rev. D11
Oct. 11, 2016
Modify “Operation Control” – “Buzzer Volume” Modify “Operation Control” – “Buzzer Tone Adjust”
Rev. D12 Rev. D13
Jan. 23, 2017 Mar. 09, 2017
Modify Operation ControlBuzzer Tone Adjust Remove Operation Control- Dollar Sign ;
Add Keyboard Interface Control- Dollar Sign Control
Rev. D14
March. 02, 2018
Add new release model A780 and A680. Modified Symbology Reading Control - Readable Symbology SettingPopular 1D Modified Getting Familiar with Your FuzzyScan- A780 and A680 series scanner
Page 5
FuzzyScan
© Copyright Cino Group © Copyright PC Worth Int’l Co., Ltd
Cino makes no warranty of any kind with regard to this publication, including, but not limited to, the implied warranty of merchantability and fitness for any particular purpose. Cino shall not be liable for errors contained herein or for incidental consequential damages in connection with the furnishing, performance, or use of this publication. This publication contains proprietary information that is protected by copyright. All rights are reserved. No part of this publication may be photocopied, reproduced or translated into any language, in any forms, in an electronic retrieval system or otherwise, without prior written permission of Cino. All product information and specifications shown in this document may be changed without prior notice.
Warranty
Cino warrants its products against defects in workmanship and materials from the date of shipment, provided that the product is operated under normal and proper conditions. The warranty provisions and durations are furnished by different warranty programs. The above warranty does not apply to any product which has been (i) misused; (ii) damaged by accident or negligence; (iii) modified or altered by the purchaser or other party; (iv) repaired or tampered by unauthorized representatives; (v) operated or stored beyond the specified operational and environmental parameters; (vi) applied software, accessories or parts are not supplied by Cino; (vii) damaged by circumstances out of Cino’s control, such as, but not limited to, lightning or fluctuation in electrical power. Any defective product must follow the warranty program and RMA procedures to return Cino for inspection.
Regulatory
Part 15 Subpart B
KN22, KN24 (KN61000-2,-3,-4,-5,-6,-8,-11)
EN55022, EN55024, EN61000-3-2, EN61000-3-3, EN60950-1, EN61000-6-3, EN61000-6-2
Class B ITE
CNS13438
LED Eye Safety
IEC62471 Exempt group
AS/NZS CISPR 22 Class B
Laser Eye Safety
IEC60825-1 Class 1
Page 6
FuzzyScan
Table of Contents
Getting Started
Getting Familiar with Your FuzzyScan 1 Connecting to Your Host 3 Using Accessories 4
Configure Your FuzzyScan
Bar Code Programming Manual 5 Programming Procedures 6 Host Interface Selection 10 Symbology Reading Control 11 Keyboard Interface Control 43 Serial Interface Control 48 Wand/Laser Emulation Control 52 Operation Control 54 Condensed DataWizard 71
Appendix
Symbology ID Table 79 Keyboard Function Code Table 81 ASCII Input Shortcut 82 Host Interface Quick Set 83 Operation Mode Quick Set 84 Option Codes 85 System Commands 87
Page 7
FuzzyScan
Getting Started
1
Getting Familiar with Your FuzzyScan
Thank you for choosing Cino FuzzyScan Bar Code Scanner. All FuzzyScan scanners deliver world-class performance for a broad range of applications to unleash your productivity.
FuzzyScan family includes A series area imager, F series linear imager and L series laser imager. The Antimicrobial models are available for A770, L780 and F780 series scanners which are equipped with Disinfectant-ready Housing and Vibrator. Moreover, the option of Vibrator is available for all other series upon request. For more details, please visit our web site or contact your supplier.
This document provides an easy reference for installation and operation purpose. The complete documentation is available at www.cino.com.tw.
A770 Series
A780 Series
A680BT Series
Power Indicator Status Indicator Trigger
Scan Window Beeper Cable Release Hole
Page 8
FuzzyScan
Getting Started
2
F790/F780/ L780 Series
F560 Series
A670/F680/L680 Series
Power Indicator Status Indicator Trigger
Scan Window Beeper Cable Release Hole
Page 9
FuzzyScan
Getting Started
3
Connecting to Your Host
FuzzyScan scanners support USB and RS-232 Serial interfaces. Please choose your desired interface cable, then plug it into the cable interface port of the scanner and connect it to the host. If you would like to remove the cable, please straighten one end of a paper clip, and then insert it into the cable release hole to pull out the cable.
RS232 Serial
USB HID & USB COM
Page 10
FuzzyScan
Getting Started
4
Using Accessories
You can enhance productivity of your workforce by using various accessories to fulfill a wide variety of application demand.
Universal Holder
The stylish Universal Holder is designed for storing your scanner when not in use. It serves to protect the scanner from lens-scratched or falling. Moreover, its artistic-design enhances the entire value of the scanner. But please note that the holder is not applicable for A series scanners.
Hand-free SmartStand
SmartStand is specifically designed for hand-free applications to maximize user’s comfort and productivity. You can adjust the scanner holder to desired position for optimized scanning.
Thanks to the auto-sense design, the scanner is capable of switching between presentation scanning and hand-held scanning automatically while working with SmartStand. But please note that this feature is not available for F500 series scanners.
In presentation mode, the barcode may not be detected by the scanner in an environment with very dim ambient lighting. You can select higher sensitivity level through the setting of Presentation Sensitivity to increase scanner’s sensitivity.
Page 11
FuzzyScan
Configure Your FuzzyScan
5
Bar Code Programming Manual
The FuzzyScan bar code commands are specially designed Proprietary bar code labels which allow you to set the FuzzyScan internal programming parameters. There are System Command, Family Code and Option Code for programming purpose.
Each programmable family and bar code command label is listed on the same page with major system commands. The detailed explanations and special programming flowchart are printed on facing or following pages. You can read the explanation and set the FuzzyScan concurrently.
A supplemental bar code command menu incorporates the bar code command labels of System Command and Option Code. As you set the FuzzyScan, open the bar code command menu to find the option code page. You may scan the desired family code and option code to set FuzzyScan. If you want to change the programming family for multiple settings, you need only turn over the programming page to find next desired programming family.
System Command
The System Command is the highest level bar code command which directs FuzzyScan to perform immediate operations, such as entering programming mode (PROGRAM), exiting programming mode (EXIT), listing system information (SYSLIST), recovering to factory preset configurations (M_DEFAULT) and so on. Please note that all system commands will take a few seconds to complete the operations. User must wait for the completion beeps before scanning another bar code.
Family Code
The Family Code is scanned to select the user desired programming family. FuzzyScan has already provided more than one hundred programming families to meet any specific requirements.
Option Code
The Option Codes is a set of bar code commands represented by “0–9”, “A–F” and finishing selection (FIN). For most setting, you must select at least one option code following the family code selection to set the desired parameter for the selected programming family.
Page 12
FuzzyScan
Configure Your FuzzyScan
6
Programming Procedures
As you scan the bar code command to select the desired parameters, information about the final selected parameters represented by the bar code commands are stored in the FuzzyScan’s internal Flash Memory ASIC or memory. If you turn off the unit, the Flash Memory ASIC or non-volatile memory retains all programming options. You need not re-program the FuzzyScan if you want to keep the existing configurations in the next power on.
The programming procedures of FuzzyScan are designed as simple as possible for ease of setting. Most programming families take the Single Scan Selection programming procedure. But several programming families have more complex and flexible programmable options, and you must take Multiple Scans Selection, Cycling Scan Selection or Dual Level Selection to complete their programming procedures. Each kind of programming procedure is listed in the following pages for your reference. Please give careful attention to become familiar with each programming procedure.
If the programming family must take multiple scans selection, cycling scan selection, or dual level selection procedures, the family of the programming menu will be marked with the matched representing symbol of Programming Category (P.C.) in bold font listed in the following table. You can easily find the bold mark in the programming menu, and refer to their flowcharts for details. Before setting the FuzzyScan, please also refer to the
“Beeping Indications” listed in Appendix to understand the
details of programming beeping indications. It will be very helpful for you to know the existing status while you are programming the FuzzyScan.
Conventions of Programming Menu
Conventions
Descriptions

Factory Default Value
P.C.
Programming Category SS
: Single scan selection MS : Multiple scans selection CS : Cycling scan selection DS : Dual level scan selection
( )
Necessary Option Code
[ ]
Selectable Option Code
Page 13
FuzzyScan
Configure Your FuzzyScan
7
Scan
END
FuzzyScan will exit programming mode, and store all
parameters in Flash Memory ASIC or non-volatile memory,
then issue the completion beeping.
Scan
PROGRAM
FuzzyScan will enter programming mode,
and inhibit all non-programming functions.
Scan
SYSLIST”
FuzzyScan will list the product information and
revision number to host via selected host interface,
then issue the completion beeping.
Scan
M_DEFAULT”
FuzzyScan will recover all programmable
parameters into factory preset configurations,
then issue the completion beeping.
Program & End
Please note that the FuzzyScan will take 3-4 seconds to store parameters in internal Flash Memory ASIC or non-volatile memory after you scan the “END”. Please don’t turn off the power before the completion beeping. It may destroy all configured parameters.
System List, Group & Master Default
Page 14
FuzzyScan
Configure Your FuzzyScan
8
Single scan selection
Multiple scans selection
Scan
PROGRAM
Scan
One of Family Codes
Scan
One of Option Codes
Repeat
Selection
Scan
END
No
Yes
Enter programming mode.
Select one of desired programming families.
Select one option code of desired parameter.
Want to select another programming family?
Exit programming mode.
Enter programming mode.
Select one of desired programming families.
1. Select one or several option codes to select desired parameters.
2. If its necessary, scan FIN to terminate option code selection.
Want to select another programming family?
Exit programming mode.
Scan
PROGRAM
Scan
One of Family Codes
Scan One or
Several Option Codes
Repeat
Selection
Scan
END
No
Yes
Page 15
FuzzyScan
Configure Your FuzzyScan
9
Cycling scan selection
Dual level selection
Enter program ming m od e .
Select one of desired prog ram m ing fam ilies.
Cyclin g s e lect o ne o r several op tion co d es o f desire d param eters as Sing le or Mu ltip le scans selectio n .
Finish cyclin g selectio n.
(If necessary)
W ant to select another prog ram m ing fam ily?
Exit program ming m od e.
Sca n
PR O GR AM
Sca n
One of Family Co des
Sca n
One of Optio n Co des
Repeat
Selection
Sca n
EN D
No
Yes
Sca n F IN
Enter program ming m od e .
Select one of desired prog ram m ing fam ilies.
Select sev e ral option co d es of d es ired param eters.
1. Select o ne o r several option co d es of d es ired para meters .
2. If its n e cess a ry, scan FIN to termina te option co d e selectio n.
W ant to select another prog ram m ing fam ily?
Exit program ming m od e.
Sca n
One of Family Co des
(1st) Sca n
Seve ral O ption C odes
Yes
Sca n
PR OG RAM
Repeat
Selection
Sca n
EN D
No
(2nd ) Scan O ne or
Seve ral O ption C odes
Page 16
FuzzyScan
Configure Your FuzzyScan
10
PROGRAM
Host Interface Selection
F_DEFAULT
Family Code Selection
P.C
Parameter Selection
Option Code
Host Interface Selection
MS MS MS MS MS MS MS MS MS MS MS MS
IBM PS/2, 25-30 series keyboard wedge interface Standard/TTL RS-232 peer-to-peer serial Wand Emulation USB Com Port Emulation PS/2 (DOS/V) direct link (keyboard replacement) PS/2 (DOS/V) keyboard wedge turbo mode PS/2 (DOS/V) keyboard wedge standard mode Laser emulation USB HID standard mode USB HID turbo mode USB HID Legacy USB EFT Terminal Mode
02 06 08 09 10 13 14 17 18 19 20 21
A series doesnt support Wand emulation, Laser emulation, USB HID Legacy and USB EFT Terminal Mode.
USB EFT Terminal Mode is only available on Trigger Mode.
Page 17
FuzzyScan
Configure Your FuzzyScan
11
PROGRAM
Symbology Reading Control
User Defined Symbol ID 
F_DEFAULT
Family Code Selection
P.C
Parameter Selection
Option Code
2nd Option Code
Symbol ID : 1 character
DS
Code 128 (default=B) GS1-128 (default=C) UPC-A (default=A) EAN-13 (default=F) Codabar/NW-7 (default=D) Code 39/Code 32 (default=G) Code 93 (default=H) Standard/Industrial 2 of 5 (default=I) Interleaved 2 of 5 (default=J) Matrix 2 of 5 (default=K) China Postal Code (default=L) German Postal Code (default=M) IATA (default=O) Code 11 (default=P) MSI/Plessey (default=R) UK/Plessey (default=S) Telepen (default=T) GS1 DataBar (default=X) UPC-E (default=E) EAN-8 (default=N) Trioptic Code 39 (default=W) UCC Coupon Extended Code (default=Z) PDF417/Micro PDF417 (default=V) Codablock F (default=Y) Code 16K (default=Q) Code 49 (default=U) Korea Post Code (default=a)
QR & Micro QR Code (default=b) Data Matrix (default=c)
Maxi Code (default=d)
00 01 02 03 04 05 06 07 08 09 10 11 12 13 14 15 16 17 18 19 20 21 22 23 24 25
26 28 29 30
(1 character) (1 character) (1 character) (1 character) (1 character) (1 character) (1 character) (1 character) (1 character) (1 character) (1 character) (1 character) (1 character) (1 character) (1 character) (1 character) (1 character) (1 character) (1 character) (1 character) (1 character) (1 character) (1 character) (1 character) (1 character) (1 character)
(1 character) (1 character) (1 character)
(1 character)
Page 18
FuzzyScan
Configure Your FuzzyScan
12
PROGRAM
Symbology Reading Control
User Defined Symbol ID 
F_DEFAULT
Family Code Selection
P.C
Parameter Selection
Option Code
2nd Option Code
Symbol ID : 1 character
DS
Aztec Code (default=e) Chinese Sensible (default=f) Australian Post (default=g) British Post (default=h) Intelligent Mail (USPS 4CB/One Code) (default=j) Japan Post (default=k) Netherlands KIX Post (default=l) US Planet (default=m) US Postnet (default=o)
31 32 33 34 36 37 38 39 41
(1 character) (1 character) (1 character) (1 character) (1 character) (1 character) (1 character) (1 character) (1 character)
Page 19
FuzzyScan
Configure Your FuzzyScan
13
PROGRAM
Symbology Reading Control
Symbology ID Transmission 
F_DEFAULT
Family Code Selection
P.C
Parameter Selection
Option Code
Symbology ID Transmission
SS SS SS SS SS SS SS
Disable symbology ID transmission

Enable prefix CINO symbology ID transmission Enable suffix CINO symbology ID transmission Enable both prefix and suffix CINO symbology ID transmission Enable prefix AIM symbology ID transmission Enable suffix AIM symbology ID transmission Enable both prefix and suffix AIM symbology ID transmission
0 1 2 3 4 5 6
Page 20
FuzzyScan
Configure Your FuzzyScan
14
PROGRAM
Symbology Reading Control
Readable Bar Code Setting 
F_DEFAULT
Family Code Selection
P.C
Parameter Selection
Option Code
Readable Symbology Setting
SS CS CS CS CS CS CS CS CS CS CS CS CS CS CS CS CS CS CS CS CS CS CS CS CS CS CS CS CS CS CS CS CS
Auto  Popular 1D Code 128 * GS1-128 * UPC-A * UPC-E * EAN-13 * EAN-8 * Codabar/NW-7 * Code 39 * Trioptic Code 39 Standard/Industrial 2 of 5 Matrix 2 of 5 Interleaved 2 of 5 * China Postal Code Germany Postal Code Code 93 * Code 11 MSI/Plessey UK/Plessey Telepen GS1 DataBar (RSS-14) * IATA PDF417 * /Micro PDF417 Codablock F Code 16K Code 49 Korea Post Code QR Code */ Micro QR Code * Data Matrix * MaxiCode Aztec Code *
Chinese Sensible (Han Xin) Code
00
C0
01 31 02 03 04 05 06 07 47 08 38 48 58 68 09 10 11 12 13 14 15 17 18 19 20
21 A0 A1 A2 A3 A4
Remember to scan FIN to terminate this selection. But if you select the Auto, FuzzyScan will terminate this selection automatically.
Page 21
FuzzyScan
Configure Your FuzzyScan
15
PROGRAM
Symbology Reading Control
Readable Bar Code Setting 
F_DEFAULT
Family Code Selection
P.C
Parameter Selection
Option Code
Readable Symbology Setting
CS CS CS CS CS CS CS
Australian Post British Post Intelligent Mail barcode Japanese Post KIX Post Planet Code Postnet
B0 B1 B3 B4 B5 B6 B8
If your application is known, you may select those known symbologies only to increase the reading speed and decrease the possibility of reading error. Furthermore, to add the Symbology ID” into the transmitted data is also helpful to identify the specific symbology.
Above symbologies marketed with * are enabled as default. When you select Auto, the scanner only reads those symbologies marked with *.
"Popular 1D" includes "Code 128", "GSA-128", "UPC-A", "UPC-E", "EAN-13", "EAN-8", "Codabar/NW-7", "Code 39", " Interleaved 2 of 5", "Code 93", "GS1 DataBar (RSS-14)".
When you set the minimum and maximum length of each symbology, please note the data length of scanned bar code doesnt include start/stop characters.
Page 22
FuzzyScan
Configure Your FuzzyScan
16
PROGRAM
Symbology Reading Control
Code 39/Code 32 Setting 
F_DEFAULT
Family Code Selection
P.C
Parameter Selection
Option Code
Code 39 Family Setting
SS SS SS SS SS SS SS SS SS SS SS SS SS SS SS
Disable Code 39 Enable Code 39 Select Standard Code 39 as primary format  Select Full ASCII Code 39 as primary format Select Code 32 (PARAF, Italian Pharmaceutical) as primary format Disable start/stop symbol transmission  Enable start/stop symbol transmission Disable Code 32 leading A transmission  Enable Code 32 leading A transmission Disable MOD 43 check digit verification  Enable MOD 43 check digit verification Disable check digit transmission  Enable check digit transmission
Disable Code 39 buffering  Enable Code 39 buffering
0 1 2 3 4 5 6 7 8 9 A B C D E
Trioptic Code 39 Setting
SS SS
Disable Trioptic Code 39

Enable Trioptic Code 39
0 1
Code 39 Min. Length
SS
MS
Default (01)

01-Maximum
Scan 2 digits from the option code chart in Appendix; then FuzzyScan will terminate this selection automatically.
FIN
(2 digits)
Code 39 Max. Length
SS
MS
Default (98)

98-Minimum
Scan 2 digits from the option code chart in Appendix; then FuzzyScan will terminate this selection automatically.
FIN
(2 digits)
Trioptic Code 39 and Code 39 Full ASCII cannot be enabled simultaneously.
Page 23
FuzzyScan
Configure Your FuzzyScan
17
PROGRAM
Symbology Reading Control
Code 39 Setting 
F_DEFAULT
Family Code Selection
P.C
Parameter Selection
Option Code
Code 39 Security Level
SS SS SS SS
Level 0 Level 1
Level 2  Level 3
0 1 2 3
Code 39 Security Level
The scanner offers four levels of decode security for Code39 bar codes: Level 0: If you are experiencing misread of poorly-printed or serious out-of-spec. bar codes in level 1, please select level 0. Level 1: If you are experiencing misread of poorly-printed or out-of-spec. bar codes in level 2, please select level 1. Level 2: This is the default setting which allows the scanner to operate fastest, while providing sufficient security in decoding in-spec Code39 bar codes. Level 3: If you failed to read poorly-printed or out-of-spec. bar codes in level 2, please select level 3. This is the most aggressive setting and may increase the misread.
Page 24
FuzzyScan
Configure Your FuzzyScan
18
PROGRAM
Symbology Reading Control
Codabar/NW-7 Setting 
F_DEFAULT
Family Code Selection
P.C
Parameter Selection
Option Code
Codabar Setting
SS SS SS SS SS SS SS SS SS SS SS SS SS SS SS
Disable Codabar Enable Codabar  Select Codabar standard format  Select Codabar ABC format Select Codabar CLSI format Select Codabar CX format Disable start/stop symbol transmission  Enable ABCD/ABCD start/stop symbol transmission Enable abcd/abcd start/stop symbol transmission Enable ABCD/TN*E start/stop symbol transmission Enable abcd/tn*e start/stop symbol transmission Disable check digit verification  Enable check digit verification Disable check digit transmission  Enable check digit transmission
0 1 2 3 4 5 6 7 8 9 A B C D E
Codabar Check Digit Settings
SS SS SS SS SS SS SS
Modulus 16

Modulus 10/weight 3 Modulus 11 Modulus 10/weight 2 7 check DR Weight Modulus 11 Runes (Modulus 10/weight 2)
0 1 2 3 4 5 6
Codabar Min. Length
SS
MS
Default (04)

01-Maximum
Scan 2 digits from the option code chart in Appendix; then FuzzyScan will terminate this selection automatically.
FIN
(2 digits)
Codabar Max. Length
SS
MS
Default (98)

98-Minimum
Scan 2 digits from the option code chart in Appendix; then FuzzyScan will terminate this selection automatically.
FIN
(2 digits)
Page 25
FuzzyScan
Configure Your FuzzyScan
19
PROGRAM
Symbology Reading Control
UPC-A & UPC-E Setting 
F_DEFAULT
Family Code Selection
P.C
Parameter Selection
Option Code
UPC Family Setting
SS SS SS SS SS SS SS SS SS SS SS SS SS SS SS SS
Disable UPC-A Enable UPC-A  Disable UPC-E Enable UPC-E  Disable UPC-E expansion

Enable UPC-E expansion Disable UPC standardization

Enable UPC standardization Disable UPC numeric system Enable UPC numeric system

Disable UPC-A check digit transmission Enable UPC-A check digit transmission

Disable UPC-E check digit transmission Enable UPC-E check digit transmission

Disable UPC “leading 1” portion

Enable UPC “leading 1” portion
0 1 2 3 4 5 6 7 8 9 A B C D E F
When enable UPC-E expansion, the UPC-E decoded data will be converted to UPC-A format and affected by related setting, such as UPC standardization, UPC numeric system, UPC-A check digit transmission.
UPC-E & EAN-8 Expansion : Expand the 8-digit UPC-E and 8-digit ENA-8 to 12-digit UPC-A and 13-digit EAN-13.
UPC-A Standardization : Expand the 12-digit UPC-A to 13-digit EAN-13 with 1 zero insertion.
UPC Lead 1 Numeric System : To read UPC leading with the 1 numeric system, you must enable this option.
WPC Selection
(UPC/EAN/CAN)
Basic Length
Disable
Check Digit
Disable
Numeric System
With 2-digit Addendum
With 5-digit Addendum
Enable
Standardization
Enable
Expansion
UPC-A
12
- 1
- 1
+ 2
+ 5
+ 1
0
UPC-E
8
- 1
- 1
+ 2
+ 5
+ 1
+ 4
EAN-13
13
- 1
NC
+ 2
+ 5
NC
0
EAN-8
8
- 1
NC
+ 2
+ 5
NC
+ 5
Page 26
FuzzyScan
Configure Your FuzzyScan
20
PROGRAM
Symbology Reading Control
UPC-A & UPC-E Setting 
F_DEFAULT
Family Code Selection
P.C
Parameter Selection
Option Code
UPC Supplement Setting
SS SS SS SS SS SS SS SS
Select UPC without supplement digits

Select UPC with only 2 supplement digits Select UPC with only 5 supplement digits Select UPC with 2/5 supplement digits Disable force supplement digits output

Enable force supplement digits output UPC Family Addenda Separator Off

UPC Family Addenda Separator On
0 1 2 3 4 5 6 7
UPC/EAN Security Level
SS SS SS
Level 0 Level 1  Level 2
Only available for UPC-A & EAN-13
0 1 2
Supplement Scan Voting
SS SS SS SS SS SS SS SS SS SS SS SS SS SS
None Level 1 Level 2 Level 3  Level 4 Level 5 Level 6 Level 7 Level 8 Level 9 Level 10 Level 11 Level 12 Level 13
0 1 2 3 4 5 6 7 8 9 A B C D
Page 27
FuzzyScan
Configure Your FuzzyScan
21
UPC/EAN Security Level
The scanner offers three levels of decode security for UPC/EAN bar codes: Level 0: If you are experiencing misread of poorly-printed or out-of-spec. bar codes, especially in characters 1, 2, 7, and 8 in level 1, please select level 0. Selection of this
security level may significantly impair the decoding ability of the scanner. Level 1: This is the default setting which allows the scanner to operate fastest, while providing sufficient security in decoding “in-spec” UPC/EAN bar codes. Level 2: If you are experiencing misread of poorly-printed, soiled or damage bar codes in level 1, please select level 2. This is the most aggressive setting and may
increase the misread.
The Supplement Scan Voting is the number of times the same UPC/EAN with 2/5 supplement digits has to be decoded before it is transmitted. It is helpful when decoding
a mix of UPC/EAN symbols with and without supplement digits. This function is effective when you select UPC/EAN with only 2 supplement digits, UPC/EAN with only 5 supplement digits or UPC/EAN with 2/5 supplement digits. The default value is Level 3. When you select higher level, it may impact the reading speed on poorly-printed, low contrast or damage barcode labels.
Page 28
FuzzyScan
Configure Your FuzzyScan
22
PROGRAM
Symbology Reading Control
EAN Setting 
F_DEFAULT
Family Code Selection
P.C
Parameter Selection
Option Code
EAN Setting
SS SS SS SS SS SS SS SS SS SS SS SS
Disable EAN-13 Enable EAN-13  Disable EAN-8 Enable EAN-8  Disable EAN-8 expansion

Enable EAN-8 expansion Disable EAN-13 check digit transmission Enable EAN-13 check digit transmission

Disable EAN-8 check digit transmission Enable EAN-8 check digit transmission

Disable ISBN/ISSN Conversion reading check

Enable ISBN/ISSN Conversion reading check
0 1 2 3 4 5 6 7 8 9 A B
EAN Supplement Setting
SS SS SS SS SS SS SS SS
Select EAN without supplement digits

Select EAN with only 2 supplement digits Select EAN with only 5 supplement digits Select EAN with 2/5 supplement digits Disable force supplement digits output

Enable force supplement digits output EAN Addenda Separator Off

EAN Addenda Separator On
0 1 2 3 4 5 6 7
Supplement Scan Voting
SS SS SS SS SS SS SS SS SS SS SS SS SS SS
None Level 1 Level 2 Level 3  Level 4 Level 5 Level 6 Level 7 Level 8 Level 9 Level 10 Level 11 Level 12 Level 13
0 1 2 3 4 5 6 7 8 9 A B C D
Page 29
FuzzyScan
Configure Your FuzzyScan
23
The Supplement Scan Voting is the number of times the same UPC/EAN with 2/5 supplement digits has to be decoded before it is transmitted. It is helpful when decoding a mix of UPC/EAN symbols with and without supplement digits. This function is effective when you select UPC/EAN with only 2 supplement digits, UPC/EAN with only 5 supplement digits or UPC/EAN with 2/5 supplement digits. The default value is Level 3. When you select higher level, it may impact the reading speed on poorly-printed, low contrast or damage barcode labels.
Page 30
FuzzyScan
Configure Your FuzzyScan
24
PROGRAM
Symbology Reading Control
EAN Setting 
F_DEFAULT
Family Code Selection
P.C
Parameter Selection
Option Code
UPC/EAN Security Level
SS SS SS
Level 0 Level 1  Level 2
Only available for UPC-A & EAN-13
0 1 2
EAN Supplement Control
SS SS SS SS SS SS SS SS
Disable all specific prefix supplement digital output  Enable all specific prefix supplement digital output Enable 491 Supplement Digit Output Enable 978/979 Supplement Digit Output Enable 977 Supplement Digit Output Enable 378/379 Supplement Digit Output Enable 414/419 Supplement Digit Output Enable 434/439 Supplement Digit Output
0 1 2 3 4 5 6 7
UPC/EAN Security Level The scanner offers three levels of decode security for UPC/EAN bar codes:
Level 0: If you are experiencing misread of poorly-printed or out-of-spec. bar codes, especially in characters 1, 2, 7, and 8 in level 1, please select level 0. Selection of this
security level may significantly impair the decoding ability of the scanner. Level 1: This is the default setting which allows the scanner to operate fastest, while providing sufficient security in decoding “in-spec” UPC/EAN bar codes. Level 2: If you are experiencing misread of poorly-printed, soiled or damage bar codes in level 1, please select level 2. This is the most aggressive setting and may
increase the misread.
EAN Supplement Control
If you select EAN with only 2, or 5 or 2/5 supplement digits and enable 491 prefix supplement digit output, the scanner will transmit EAN with 2, or 5 or 2/5 supple ment
digits bar codes starting with 491 prefix. The EAN without supplement digit will not be transmitted. If you select EAN with only 2, or 5 or 2/5 supplement digits and enable the other except 491 prefix supplement digit output, the scanner will transmit EAN with 2, or 5, or 2/5 supplement digits bar codes starting with specific prefix. The EAN without supplement digit will be transmitted.
Page 31
FuzzyScan
Configure Your FuzzyScan
25
PROGRAM
Symbology Reading Control
UCC Coupon Extended Code Setting 
F_DEFAULT
Family Code Selection
P.C
Parameter Selection
Option Code
UCC Coupon Extended Code
SS SS
Disable UCC Coupon Extended Code  Enable UCC Coupon Extended Code
0 1
UCC Coupon Extended Code When UCC coupon extended code function is enabled, scanner decodes UPC-A barcodes starting with digit 5, EAN-13 barcodes starting with digit 99 and GS1-128 Coupon Codes. UPC-A, EAN-13 and EAN-128 must be enabled to scan all types of Coupon Codes.
Page 32
FuzzyScan
Configure Your FuzzyScan
26
PROGRAM
Symbology Reading Control
IATA & Interleaved 2 of 5 Setting 
F_DEFAULT
Family Code Selection
P.C
Parameter Selection
Option Code
IATA Setting
SS SS SS SS SS SS SS SS SS SS SS SS SS
Disable IATA  Enable IATA Select 15-digit fixed length IATA checking  Select variable length IATA Disable check digit verification  Enable check digit automatic verification Enable S/N checking digit verification only Enable CPN checking digit verification only Enable CPN, Airline and S/N check digit verification Disable check digit transmission  Enable check digit transmission Disable start/stop symbol transmission

Enable start/stop symbol transmission
0 1 2 3 4 5 6 7 8 9 A B C
Interleaved 2 of 5 Setting
SS SS SS SS SS SS SS SS SS
Disable Interleaved 2 of 5 Enable Interleaved 2 of 5  Select Interleaved 2 of 5 as primary format  Select German Postal Code as primary format No check character  Validate USS check digit Validate OPCC check digit Disable check digit transmission

Enable check digit transmission
0 1 2 3 4 5 6 7 8
Page 33
FuzzyScan
Configure Your FuzzyScan
27
PROGRAM
Symbology Reading Control
Code 25 Family Setting 
F_DEFAULT
Family Code Selection
P.C
Parameter Selection
Option Code
Code 25 Setting
SS SS SS SS SS SS SS SS SS SS
Disable Standard/Industrial 2 of 5  Enable Standard/Industrial 2 of 5 Disable Matrix 2 of 5  Enable Matrix 2 of 5 Disable China Postal Code  Enable China Postal Code Disable check digit verification

Enable check digit verification Disable check digit transmission  Enable check digit transmission
0 1 2 3 4 5 6 7 8 9
Code 25 Family Min. Length
SS
MS
Default (04)

01-Maximum
Scan 2 digits from the option code chart in Appendix; then FuzzyScan will terminate this selection automatically.
FIN
(2 digits)
Code 25 Family Max. Length
SS
MS
Default (98)

98-Minimum
Scan 2 digits from the option code chart in Appendix; then FuzzyScan will terminate this selection automatically.
FIN
(2 digits)
For Code25 setting, we recommend you to select only one type of Code 25 or set the maximum/minimum bar code length. To decode all types of Code 25 or to variable length of Code 25 will increase the possibility of reading error.
Page 34
FuzzyScan
Configure Your FuzzyScan
28
PROGRAM
Symbology Reading Control
Code 11 & Code 93 Setting 
F_DEFAULT
Family Code Selection
P.C
Parameter Selection
Option Code
Code 11 Setting
SS SS SS SS SS SS SS
Disable Code 11  Enable Code 11 Disable check digit verification  Select 1-check digit verification Select 2-check digit verification Disable check digit transmission

Enable check digit transmission
0 1 2 3 4 5 6
Code 11 Min. Length
SS
MS
Default (04)

01-Maximum
Scan 2 digits from the option code chart in Appendix; then FuzzyScan will terminate this selection automatically.
FIN
(2 digits)
Code 11 Max. Length
SS
MS
Default (98)

98-Minimum
Scan 2 digits from the option code chart in Appendix; then FuzzyScan will terminate this selection automatically.
FIN
(2 digits)
Code 93 Setting
SS SS SS SS
Disable Code 93 Enable Code 93  Disable check digit transmission  Enable check digit transmission
0 1 2 3
Code 93 Min. Length
SS
MS
Default (01)

01-Maximum
Scan 2 digits from the option code chart in Appendix; then FuzzyScan will terminate this selection automatically.
FIN
(2 digits)
Code 93 Max. Length
SS
MS
Default (98)

98-Minimum
Scan 2 digits from the option code chart in Appendix; then FuzzyScan will terminate this selection automatically.
FIN
(2 digits)
Page 35
FuzzyScan
Configure Your FuzzyScan
29
PROGRAM
Symbology Reading Control
MSI/Plessey Setting 
F_DEFAULT
Family Code Selection
P.C
Parameter Selection
Option Code
MSI/Plessey Setting
SS SS SS SS SS SS SS
Disable MSI/Plessy  Enable MSI/Plessy Select MOD 10 check digit

Select MOD 10-10 check digit Select MOD 11-10 check digit Disable check digit transmission  Enable check digit transmission
0 1 2 3 4 5 6
MSI/Plessey Min. Length
SS
MS
Default (04)

01-Maximum
Scan 2 digits from the option code chart in Appendix; then FuzzyScan will terminate this selection automatically.
FIN
(2 digits)
MSI/Plessey Max. Length
SS
MS
Default (98)

98-Minimum
Scan 2 digits from the option code chart in Appendix; then FuzzyScan will terminate this selection automatically.
FIN
(2 digits)
Page 36
FuzzyScan
Configure Your FuzzyScan
30
PROGRAM
Symbology Reading Control
Code 128 Setting 
F_DEFAULT
Family Code Selection
P.C
Parameter Selection
Option Code
Code 128 Setting
SS SS SS SS
Disable Code 128 Enable Code 128  ISBT Concatenation Off

ISBT Concatenation On
0 1 2 3
Code 128 Min. Length
SS
MS
Default (01)

01-Maximum
Scan 2 digits from the option code chart in Appendix; then FuzzyScan will terminate this selection automatically.
FIN
(2 digits)
Code 128 Max. Length
SS
MS
Default (98)

98-Minimum
Scan 2 digits from the option code chart in Appendix; then FuzzyScan will terminate this selection automatically.
FIN
(2 digits)
Code 128 Security Level
SS SS
Level 0 Level 1  0 1
Code 128 Security Level The scanner offers two levels of decode security for Code128 bar codes: Level 0: If you are experiencing misread of poor-printed or out-of-spec. bar code in level1, please select level 0. Level 1: This is the default setting which allows the scanner to operate fastest, while providing sufficient security in decoding "in-spec." Code128 bar codes.
Page 37
FuzzyScan
Configure Your FuzzyScan
31
PROGRAM
Symbology Reading Control
GS1-128 Setting 
F_DEFAULT
Family Code Selection
P.C
Parameter Selection
Option Code
GS1-128 Setting
SS SS
Disable GS1-128 Enable GS1-128 
0 1
GS1-128 Min. Length
SS
MS
Default (01)  01-Maximum
Scan 2 digits from the option code chart in Appendix; then FuzzyScan will terminate this selection automatically.
FIN
(2 digits)
GS1-128 Max. Length
SS
MS
Default (98)  98-Minimum
Scan 2 digits from the option code chart in Appendix; then FuzzyScan will terminate this selection automatically.
FIN
(2 digits)
Page 38
FuzzyScan
Configure Your FuzzyScan
32
PROGRAM
Symbology Reading Control
UK/Plessey Setting 
F_DEFAULT
Family Code Selection
P.C
Parameter Selection
Option Code
UK/Plessey Setting
SS SS SS SS SS SS SS SS
Disable UK/Plessey  Enable UK/Plessey Select UK/Plessey Standard Format

Select UK/Plessey CLSI Format Disable Convert X to A-F

Enable Convert X to A-F Disable check digit transmission

Enable check digit transmission
0 1 2 3 4 5 6 7
UK/Plessey Min. Length
SS
MS
Default (04)

01-Maximum
Scan 2 digits from the option code chart in Appendix, then FuzzyScan will terminate this selection automatically.
FIN
(2 digits)
UK/Plessey Max. Length
SS
MS
Default (98)

98-Minimum
Scan 2 digits from the option code chart in Appendix, then FuzzyScan will terminate this selection automatically.
FIN
(2 digits)
Page 39
FuzzyScan
Configure Your FuzzyScan
33
PROGRAM
Symbology Reading Control
Telepen Setting 
F_DEFAULT
Family Code Selection
P.C
Parameter Selection
Option Code
Telepen Setting
SS SS SS SS SS SS
Disable Telepen  Enable Telepen Select Telepen Numeric mode

Select Telepen Full ASCII mode Disable check digit transmission

Enable check digit transmission
0 1 2 3 4 5
Telepen Min. Length
SS
MS
Default (04)

01-Maximum
Scan 2 digits from the option code chart in Appendix, then FuzzyScan will terminate this selection automatically.
FIN
(2 digits)
Telepen Max. Length
SS
MS
Default (98)

98-Minimum
Scan 2 digits from the option code chart in Appendix, then FuzzyScan will terminate this selection automatically.
FIN
(2 digits)
Page 40
FuzzyScan
Configure Your FuzzyScan
34
PROGRAM
Symbology Reading Control
GS1 DataBar Setting 
F_DEFAULT
Family Code Selection
P.C
Parameter Selection
Option Code
GS1 DataBar Setting
SS SS SS SS SS SS
Disable GS1 DataBar (RSS-14) Enable GS1 DataBar (RSS-14)  Disable GS1 DataBar Limited Enable GS1 DataBar Limited  Disable GS1 DataBar Expanded Enable GS1 DataBar Expanded 
0 1 2 3 4 5
GS1 DataBar Limited
Security Level
SS SS SS
Level 1 Level 2 Level 3

Only available for GS1 DataBar Limited
Only available for F460, F560 scanners.
0 1 2
GS1 DataBar Min. Length
SS
MS
Default (04)

01-Maximum
Only available for GS1 DataBar Expanded Scan 2 digits from the option code chart in Appendix; then FuzzyScan will terminate
this selection automatically.
FIN
(2 digits)
GS1 DataBar Max. Length
SS
MS
Default (74)

74-Minimum
Only available for GS1 DataBar. Expanded Scan 2 digits from the option code chart in Appendix; then FuzzyScan will terminate this selection automatically.
FIN
(2 digits)
GS1 128 Limited Security Level The scanner F460/F560 offers three levels of decode security for GS1 DataBar Limited bar codes:
Level 1: If you failed to read poorly-printed or out-of-spec. bar codes in level 2, please select level 1. This is the most aggressive setting and may increase the misread. Level 2: If you are experiencing misread of poor-printed or out-of-spec. bar code in level 3, please select level 2. Level 3: This is the default setting which allows the scanner to operate fastest, while providing sufficient security in decoding "in-spec." GS1 128 Limited bar codes.
Page 41
FuzzyScan
Configure Your FuzzyScan
35
PROGRAM
Symbology Reading Control
Composite Codes, Codablock F
PDF417/MicroPDF417 & Setting
F_DEFAULT
Family Code Selection
P.C
Parameter Selection
Option Code
Composite Codes Setting
SS SS SS SS
Disable composite codes  Enable composite codes UPC Composite Mode: UPC never linked  UPC Composite Mode: UPC always linked
If UPC Composite Mode: UPC never linked is selected, UPC barcodes are transmitted whether MicroPDF417 symbol is detected or not.
If UPC Composite Mode: UPC always linked is selected, UPC barcodes are only transmitted when the MicroPDF417 is detected.
0 1 2 3
Codablock F Setting
SS SS
Disable

Enable
0 1
PDF417/Micro PDF417 Setting
SS SS SS SS
Disable PDF417 Enable PDF417  Disable MicroPDF417  Enable MicroPDF417
0 1 2 3
Page 42
FuzzyScan
Configure Your FuzzyScan
36
PROGRAM
Symbology Reading Control
Code 16K & Code 49 Setting 
F_DEFAULT
Family Code Selection
P.C
Parameter Selection
Option Code
Code 16K Setting
SS SS
Disable Code 16K  Enable Code 16K 0 1
Code 16K Min. Length
SS
MS
Default (01)

01-Maximum
Scan 3 digits from the option code chart in Appendix; then FuzzyScan will terminate this selection automatically.
FIN
(3 digits)
Code 16K Max. Length
SS
MS
Default (160)

160-Minimum
Scan 3 digits from the option code chart in Appendix; then FuzzyScan will terminate this selection automatically.
FIN
(3 digits)
Code 49 Setting
SS SS
Disable Code 49  Enable Code 49 0 1
Code 49 Min. Length
SS
MS
Default (01)

01-Maximum
Scan 2 digits from the option code chart in Appendix; then FuzzyScan will terminate this selection automatically.
FIN
(2 digits)
Code 49 Max. Length
SS
MS
Default (81)

81-Minimum
Scan 2 digits from the option code chart in Appendix; then FuzzyScan will terminate this selection automatically.
FIN
(2 digits)
Page 43
FuzzyScan
Configure Your FuzzyScan
37
PROGRAM
Symbology Reading Control
QR Code Setting 
F_DEFAULT
Family Code Selection
P.C
Parameter Selection
Option Code
QR Code Setting
SS SS
Disable QR Code
Enable QR Code  Disable MicroQR Code
Enable MicroQR Code  Disable QR Code Append Enable QR Code Append  Disable QR Code Inverse Reading Enable QR Code Inverse Reading Auto detect QR Code Inverse Reading
0 1 2 3
4 5
6 7 8
QR Code Min. Length
SS
MS
Default (01)

01-Maximum
Scan 4 digits from the option code chart in Appendix; then FuzzyScan will terminate this selection automatically.
FIN
(4 digits)
QR Code Max. Length
SS
MS
Default (7089)

7089-Minimum
Scan 4 digits from the option code chart in Appendix; then FuzzyScan will terminate this selection automatically.
FIN
(4 digits)
Page 44
FuzzyScan
Configure Your FuzzyScan
38
PROGRAM
Symbology Reading Control
Data Matrix Setting 
F_DEFAULT
Family Code Selection
P.C
Parameter Selection
Option Code
Data Matrix Setting
SS SS SS SS SS SS SS SS
Disable Data Matrix Enable Data Matrix  Disable Data Matrix Inverse Reading Enable Data Matrix Inverse Reading Auto Detect Data Matrix Inverse Reading  Disable Data Matrix Mirror Images Enable Data Matrix Mirror Images Auto Detect Data Matrix Mirror Images 
0 1 4 5 6 7 8 9
Data Matrix Min. Length
SS
MS
Default (01)

01-Maximum
Scan 4 digits from the option code chart in Appendix; then FuzzyScan will terminate this selection automatically.
FIN
(4 digits)
Data Matrix Max. Length
SS
MS
Default (3116)

3116-Minimum
Scan 4 digits from the option code chart in Appendix; then FuzzyScan will terminate this selection automatically.
FIN
(4 digits)
Small DM Code Reading
SS SS SS
Standard

Level 1 Level 2
0 1 2
Small DM Code Reading:
When small DataMatrix code can’t be read by 2D scanner, you can select "Level 1” or “Level 2” to improve the scanner's ability to read small DataMatrix code. The scanner’s snappiness decreased when you select “Level 1” or “Level 2”. The higher level will take longer time to read the small DataMatrix barcode.
Small DM Code Reading: Available firmware: A780 / A680 1.00.01 and above
A770 1.00.24 and above; 2.00.08 and above A670 1.00.05 and above
Page 45
FuzzyScan
Configure Your FuzzyScan
39
PROGRAM
Symbology Reading Control
MaxiCode Setting 
F_DEFAULT
Family Code Selection
P.C
Parameter Selection
Option Code
MaxiCode Setting
SS
SS
Disable MaxiCode 
Enable MaxiCode 0 1
MaxiCode Min. Length
SS
MS
Default (01)

01-Maximum
Scan 3 digits from the option code chart in Appendix, then FuzzyScan will terminate this selection automatically.
FIN
(3 digits)
MaxiCode Max. Length
SS
MS
Default (150)

150-Minimum
Scan 3 digits from the option code chart in Appendix, then FuzzyScan will terminate this selection automatically.
FIN
(3 digits)
Page 46
FuzzyScan
Configure Your FuzzyScan
40
PROGRAM
Symbology Reading Control
Aztec Code Setting 
F_DEFAULT
Family Code Selection
P.C
Parameter Selection
Option Code
Aztec Code Setting
SS SS
Disable Aztec Code Enable Aztec Code 
0 1
Aztec Code Min. Length
SS
MS
Default (01)  01-Maximum
Scan 4 digits from the option code chart in Appendix; then FuzzyScan will terminate this selection automatically.
FIN
(4 digits)
Aztec Code Max. Length
SS
MS
Default (3832)  3832-Minimum
Scan 4 digits from the option code chart in Appendix; then FuzzyScan will terminate this selection automatically.
FIN
(4 digits)
Page 47
FuzzyScan
Configure Your FuzzyScan
41
PROGRAM
Symbology Reading Control
Australian Post, US Planet, US Postnet,
British Post & Japan Post Setting 
F_DEFAULT
Family Code Selection
P.C
Parameter Selection
Option Code
Australian Post Setting
SS SS SS SS SS SS
Disable Australian Post  Enable Australian Post Raw format Output  Numeric Encoding Output (N Encoding Table) Alphanumeric Encoding Output (C Encoding Table) Auto-discriminate Output (Combination C & N Encoding Table)
0 1 2 3 4 5
US Planet Setting
SS SS SS SS
Disable US Planet  Enable US Planet Disable Check Digit Transmission  Enable Check Digit Transmission
0 1 2 3
US Postnet Setting
SS SS SS SS
Disable US Postnet  Enable US Postnet Disable Check Digit Transmission  Enable Check Digit Transmission
0 1 2 3
British Post Setting
SS SS SS SS
Disable British Post  Enable British Post Disable Check Digit Transmission  Enable Check Digit Transmission
0 1 2 3
Japan Post Setting
SS SS
Disable Japan Post  Enable Japan Post 0 1
Australian Post Setting: Auto-discriminate output option increase the risk of misread because the encoded data format does not specify the Encoding Table used for
encoding.
Page 48
FuzzyScan
Configure Your FuzzyScan
42
PROGRAM
Symbology Reading Control
Netherlands KIX Code, Intelligent Mail
& Korea Post Code Setting 
F_DEFAULT
Family Code Selection
P.C
Parameter Selection
Option Code
Netherlands KIX Code Setting
SS SS
Disable Netherlands KIX Code

Enable Netherlands KIX Code
0 1
Intelligent Mail Setting (USPS 4CB/One Code)
SS SS
Disable Intelligent Mail  Enable Intelligent Mail
0 1
Korea Post Code Setting
SS SS
Disable  Enable Length fixed in 6 characters.
0 1
Page 49
FuzzyScan
Configure Your FuzzyScan
43
PROGRAM
Keyboard Interface Control
Keyboard Layout (Language) Setting 
F_DEFAULT
Family Code Selection
P.C
Parameter Selection
Option Code
Keyboard Layout
SS SS SS SS SS SS SS SS SS SS SS SS SS SS SS SS SS SS
USA (QWERTY)

France (AZERTY) Germany (QWERTZ) United Kingdom - UK (QWERTY) Canadian French (QWERTY) Spain (Spanish, QWERTY) Sweden/Finland (QWERTY) Portugal (QWERTY) Norway (QWERTY) Spain (Latin America, QWERTY) Italy (QWERTY) Netherlands (QWERTY) Denmark (QWERTY) Belgium (AZERTY) Switzerland-Germany (QWERTZ) Iceland (QWERTY) Japan (DOS/V) Czech (QWERTY)
00 01 02 03 04 05 06 07 08 09 10 11 12 13 14 15 16 17
Please refer to the ASCII/HEX Table listed in the Appendix to determine HEX codes for characters, symbols, and functions to be used as preamble or postamble.
To set preamble or postamble as function key output, you must enable the “Function Key Emulation” feature as listed in page 3-25 first.
Keyboard Interface Message String :
Preamble
Data Length
Prefix Symbol ID
Scanned Data
Suffix Symbol ID
Postamble
Record Suffix
1-15 characters
2-4 digits
1 or 3 characters
Variable length
1 or 3 characters
1-15 characters
1 character
Page 50
FuzzyScan
Configure Your FuzzyScan
44
PROGRAM
Keyboard Interface Control
Record Suffix, Preamble, Postamble,
FNC1 Transmit & Caps Lock 
F_DEFAULT
Family Code Selection
P.C
Parameter Selection
Option Code
Record Suffix
SS SS SS SS SS SS
None RETURN

TAB SPACE ENTER (Numeric Key Pad) User defined character (1 character)
0 1 2 3 4
5, (00-7F)
Preamble
SS
MS
None

1-15 characters
Maximum 15-character input; scan “FIN” to terminate this selection.
FIN
[00-7F], [FIN]
Postamble
SS
MS
None

1-15 characters
Maximum 15-character input; scan “FIN” to terminate this selection.
FIN
[00-7F], [FIN]
FNC1 Symbol Char. Transmit
SS SS
Disable Enable

0 1
Caps Lock Control
SS SS SS
“Caps Lock Off” State

“Caps Lock On” State Auto Detect (PC/AT, PS/2, Keyboard Replacement and DOS/V Machines only)
0 1 2
Caps Lock Release Control
SS SS
Caps Lock On, Caps Off” 
Caps Lock On, Shift Off
0 1
Page 51
FuzzyScan
Configure Your FuzzyScan
45
FNC1 Symbol Char. Transmit: When this function is enabled and the FNC1 is contained in scanned data, the scanner transmits the FNC1 to the host. Chart of theFNC1 is provided in Appendix- Keyboard Function Code Table. When the scanner interface is set to keyboard, the scan code is converted to corresponding key function before it is transmitted.
The function of “Caps Lock Control” and “Key Pad Emulation” are only available for IBM PC/AT, PS/VP, PS/2 series personal computers and compatible machines. While selecting the other host interfaces, these selections don’t perform the above functions for you.
Please check the actual Caps Lock state in use while software application is running. If the Caps Lock state is off, select “Caps Lock Off” state, and then FuzzyScan will perform normal data transmission. If the Caps Lock state is on, select “Caps Lock On” state. Select “Auto Detect”, FuzzyScan will perform special transmission
handshaking without changing the status of Caps Lock switch.
Page 52
FuzzyScan
Configure Your FuzzyScan
46
PROGRAM
Keyboard Interface Control
Delay Setting 
F_DEFAULT
Family Code Selection
P.C
Parameter Selection
Option Code
Intermessage Delay
SS
MS
None

1-99 (x5) msec.
Scan 2 digits from the option code chart in Appendix; then FuzzyScan will terminate this selection automatically.
FIN
(2 digits)
Intercharacter Delay
SS
MS
None

1-99 (x5) msec.
Scan 2 digits from the option code chart in Appendix; then FuzzyScan will terminate this selection automatically.
FIN
(2 digits)
Interfunction Delay
SS
MS
None

1-99 (x5) msec. Scan 2 digits from the option code chart in Appendix; then FuzzyScan will terminate
this selection automatically.
FIN
(2 digits)
Intermessage Delay is a time delay between messages output by FuzzyScan. Increasing this delay will help host applications process the incoming data on time.
Intercharacter Delay is a time delay between data characters output by FuzzyScan. These two parameters are used to synchronize data communication when : 1) the data
transmission speed is too fast, characters may be skipped; 2) multitasking operation system or host computers in a network may slow down the keyboard handling; 3)
various notebook or desktop PC systems require different timing parameter settings. Please always add one extra unit as safety margin when adjusting these two
parameters.
Interfunction Delay is a time delay of transmission of segments in each message string.
Page 53
FuzzyScan
Configure Your FuzzyScan
47
PROGRAM
Keyboard Interface Control
Emulation Setting, Key Pad Emulation &
Upper/Lower Case Setting , Dollar Sign Control
F_DEFAULT
Family Code Selection
P.C
Parameter Selection
Option Code
Function Key Emulation
SS SS
Enable ASCII 00-31 code as keyboard function code output

Ctrl-Output Refer to Appendix – Keyboard Function Code Table for details.
0 1
Key Pad Emulation
SS SS
Disable key pad emulation

Enable numeric output as key pad (Num Lock On) output
0 1
Upper/Lower Case
SS SS SS SS
Normal case (neglect the upper/lower case control)

Inverse case (change all characters output to inverse case) Upper case (force all characters output as upper case) Lower case (force all characters output as lower case)
0 1 2 3
Dollar Sign Control
SS SS SS SS SS
Dollar sign output as “ $ “  Dollar sign output as “¥“ Dollar sign output as “ € “ Dollar sign output as “ £ “ Dollar sign output as “ ¢ “
0 1 2 3 4
Page 54
FuzzyScan
Configure Your FuzzyScan
48
PROGRAM
Serial Interface Control
Record Suffix, Preamble ,Postamble Setting 
F_DEFAULT
Family Code Selection
P.C
Parameter Selection
Option Code
STX/ETX Control
SS SS
Disable STX/ETX transmission

Enable STX/ETX transmission STX/ETX are two characters used to indicate the starting and ending of the total
data frame transmitted via serial interface.
0 1
Record Suffix
SS SS SS SS SS SS
None CR (0DH)

LF (0AH) CRLF (0D0AH) TAB (09H) SPACE (20H)
0 1 2 3 4 5
Preamble
SS
MS
None

1-15 characters Maximum 15-character input; scan “FIN” to terminate this selection.
FIN
[00-7F], [FIN]
Postamble
SS
MS
None

1-15 characters
Maximum 15-character input; scan “FIN” to terminate this selection.
FIN
[00-7F], [FIN]
FNC1 Symbol Char. Transmit
SS SS
Disable Enable

When this function is enabled and the FNC1 is contained in scanned data, the scanner transmits the FNC1 to the host. Chart of the FNC1 is provided in Appendix – Keyboard Function Code Table. When the scanner interface is set to keyboard, the scan code is converted to corresponding key function before it is transmitted
0 1
Serial Interface Message String (RS232, USB COM) :
STX
Preamble
Data Length
Prefix Symbol ID
Scanned Data
Suffix Symbol ID
Postamble
ETX
Record Suffix
1 character
1-15 characters
2-4 digits
1 or 3 characters
Variable length
1 or 3 characters
1-15 characters
1 character
1 character
Page 55
FuzzyScan
Configure Your FuzzyScan
49
PROGRAM
Serial Interface Control
Delay Setting 
F_DEFAULT
Family Code Selection
P.C
Parameter Selection
Option Code
Intermessage Delay
SS
MS
None

1-99 (x5) msec.
Scan 2 digits from the option code chart in Appendix; then FuzzyScan will terminate this selection automatically.
FIN
(2 digits)
Intercharacter Delay
SS
MS
None

1-99 (x5) msec.
Scan 2 digits from the option code chart in Appendix; then FuzzyScan will terminate this selection automatically.
FIN
(2 digits)
Interfunction Delay
SS
MS
None

1-99 (x5) msec.
Scan 2 digits from the option code chart in Appendix; then FuzzyScan will terminate this selection automatically.
FIN
(2 digits)
Intermessage Delay is a time delay between messages output by FuzzyScan. Increasing this delay will help host applications process the incoming data on time.
Intercharacter Delay is a time delay between data characters output by FuzzyScan. These two parameters are used to synchronize data communication when : 1) the data
transmission speed is too fast, characters may be skipped; 2) multitasking operation system or host computers in a network may slow down the keyboard handling; 3)
various notebook or desktop PC systems require different timing parameter settings. Please always add one extra unit as safety margin when adjusting these two
parameters.
Interfunction Delay is a time delay between transmission and reception of each segment of the message string.
Page 56
FuzzyScan
Configure Your FuzzyScan
50
PROGRAM
Serial Interface Control
Protocol, ACK/NAK Setting 
F_DEFAULT
Family Code Selection
P.C
Parameter Selection
Option Code
Handshaking Protocol
SS SS SS SS
None (free running mode)

RTS/CTS (hardware handshaking) ACK/NAK (software handshaking) Xon/Xoff (software handshaking)
0 1 2 3
NAK Retry Count
SS SS
3 times  0~255 times
FIN
(3 digits)
ACK Indication
SS SS SS SS
Disable ACK Time-out Indication Enable ACK Time-out Indication

Disable ACK Indication  Enable ACK Indication
0 1 2 3
USB COM doesnt support RTS/CTS handshaking protocol.
When the ACK/NAK Software Handshaking option is selected, the FuzzyScan waits for an ACK (acknowledge) or NAK (not acknowledge) from the host computer after each data transmission. If the NAK is received, FuzzyScan will re-send the data until receiving ACK.
NAK Retry Count
After transmitting data, the scanner expects a NAK response from the host up to the preset Serial Response Time-out. If the scanner doesnt get a response, the scanner
will issue an error indication and discard the data. When a NAK is received, the scanner transmits the same data again and waits for either an ACK or NAK. The scanner
issues an error indication and discards the data under following two conditions:
1) After preset NAK retry counts is received within the preset serial response time-out.
2) If the preset time-out is up but the preset NAK retry counts havent come to the end.
The default retry counts are three times. If you program 0 time, the scanner wont resend the data to the host when the scanner receives a NAK. The scanner will discard
the data. If you program “255 times”, the scanner can receive unlimited NAKs from the host within the pre-set serial response time-out.
This function is not available for batch mode. When you enable this function in on-line mode, the out-of-range function will be disable automatically.
ACK Indication:
Disable: Theres neither LED nor beeping indication for this setting.
Enable: Theres a specific LED and beeping indication for this setting.
Page 57
FuzzyScan
Configure Your FuzzyScan
51
PROGRAM
Serial Interface Control
Response Time-out Setting, Baud Rate, Data Frame
F_DEFAULT
Family Code Selection
P.C
Parameter Selection
Option Code
Serial Response Time-out
SS SS SS SS SS SS
None 200 mseconds 500 mseconds

800 mseconds 1 second 2 seconds
3 seconds 4 seconds 5 seconds 8 seconds 10 seconds 15 seconds
0 1 2 3 4 5
6 7 8
9 A B
Baud Rate (BPS)
SS SS SS SS
38.4K BPS
19.2K BPS 9600 BPS

4800 BPS
2400 BPS 1200 BPS
57.6K BPS
115.2K BPS
0 1 2
3
4
5
8
9
Data Frame
SS SS SS SS SS SS SS SS
8, None, 1

8, Odd, 1 8, Even, 1 8, Space, 1 8, Mark, 1 8, None, 2 7, Odd, 1 7, Even, 1
7, Space, 1 7, Mark, 1 7, None, 2 7, Odd, 2 7, Even, 2 7, Space, 2 7, Mark, 2
0 1 2 3 4 5 6 7
8
9 A B C D E
When the RTS/CTS Hardware Handshaking option is selected, the RTS (request to send) and CTS (clear to send) signals will be issued before normal data communication. This option is very helpful to ensure the reliability of data communication.
The Serial Response Time-out is a pre-defined delay time for FuzzyScan to wait for handshaking, acknowledgment or non-acknowledgment from the host computer
Page 58
FuzzyScan
Configure Your FuzzyScan
52
PROGRAM
Wand/Laser Emulation Control
(F & L Series)
Output Polarity, Signal State, Margin/Module Time 
F_DEFAULT
Family Code Selection
P.C
Parameter Selection
Option Code
Output Polarity
SS SS
High level (5Vdc) on Bar (low level on Space)

Low level (0Vdc) on Bar (high level on Space)
Determine the output voltage level for both bar and space.
0 1
Initial Signal State
SS SS
High Level (5Vdc)

Low Level (0Vdc)
Determine the initial state of output voltage level.
0 1
Margin Time
SS SS SS SS SS
10 msec. 15 msec. 20 msec.

25 msec. 30 msec.
0 1 2 3 4
Module Time
SS SS SS SS
Extremely short Short Medium  Long
0 1 2 3
Narrow/Wide Ratio
SS SS SS
1:2

1:2.5 1:3
0 1 2
Page 59
FuzzyScan
Configure Your FuzzyScan
53
PROGRAM
Wand/Laser Emulation Control
(F & L Series)
Output Polarity, Signal State, Margin/Module Time 
F_DEFAULT
Family Code Selection
P.C
Parameter Selection
Option Code
Code 39/Code 128 Emulation
SS SS SS SS SS
Disable standard Code 39 emulation  Enable standard Code 39 skip emulation Enable standard Code 39 replace emulation Enable Full ASCII Code 39 emulation Enable Code 128 emulation
0 1 2 3 4
[ Code 39 Skip] : When this option is selected, all scanned data will be translated as Standard Code 39 wand/laser emulation output. If any lower case characters are read,
they will be translated to upper case characters. Any other characters that are not available in Code 39 symbology set will be skipped.
[ Code 39 Replace] : Any character not normally available in the standard Code 39 symbology set, will be translated as Space.
Page 60
FuzzyScan
Configure Your FuzzyScan
54
PROGRAM
Operation Control
(F & L Series)
Operation Mode Setting 
F_DEFAULT
Family Code Selection
P.C
Parameter Selection
Option Code
Operation Mode
SS SS SS SS SS SS SS SS SS
Low Power mode Trigger mode

Presentation mode Alternative mode Flash mode Force mode Toggle mode Diagnostic mode Level mode
0 1 2 3 4 5 6 7 8
Low Power Mode (Low Power Triggering): The scanner goes into idle state after scanning the bar code. You must press the trigger to wake up the scanner for operation.
It is very helpful for mobile data collection and application, which are concerned with power saving.
Trigger Mode (External Triggering): The scanner goes into standby state after scanning the bar code. You must press the trigger to turn on the light source of the scanner
before scanning the bar code.
Presentation Mode (Auto Detection): Presentation mode uses ambient light to detect the bar codes. The light source is off until the scanner detects an image which is
similar to a barcode. Then the light source turns on automatically to read the bar code. If the light level in the room is not high enough, Presentation Mode may not work properly. You can choose different level of Presentation Sensitivity to meet your application (Please refer to the setting of Presentation Sensitivity).
Alternative Mode (Periodic Power Off) :The scanner keeps the light source of the scanner turned on till the pre-defined light source on time is up. After the scanner turns
off the light source, you must press the trigger to turn on the light source again. After each good read, the timer counter of Light Source on Time is reset. For you do not have to press the trigger frequently, it is very convenient for multiple scanning.
Flash Mode (Pulse Driven Reading): The scanner flashes the light source of the scanner without using the trigger. If the scanner detects an image which is similar to a bar
code, the scanner forces on the light source automatically and scans the bar code. Flash Duty Cycle adjustment can change the frequency of the blinking.
Force Mode (Continued Power On): The light source of the scanner is forced on for continued operation without pressing the trigger switch. This mode is convenient for
high speed bar code reading.
Toggle Mode (Repeat Reading): The toggle mode is very similar to the Alternative Mode but without the pre-defined light source on time concern. You must press the
trigger to turn on the light source of the scanner to scan. The scanner keeps the light source turned on until you press the trigger again.
Diagnostic Mode (Test Reading): This operation mode is specifically designed for diagnostic purposes. When this operation mode is selected, the light source of the
scanner is force on without regard for other programmable parameters, such as reread delay, redundancy, and so forth.
Level Mode (Auto Power Off): When this operation mode is selected, the scanner continues to turn on the light source of the scanner before a good read or pre-defined
Light Source on Time. If the scanner decodes a bar code successfully, it turns off the light source immediately. After the scanner turns off the light source, you must press the trigger to turn on the light source again. If there is no scanning operation performed during the pre -defined light source on time, the scanner enters the idle state after the pre-defined light source on time is up.
FuzzyScan Laser model only have LED illumination (without laser aiming line) in Flash/ Force/Toggle/ Diagnostics Modes.
Page 61
FuzzyScan
Configure Your FuzzyScan
55
PROGRAM
Operation Control
(F & L Series)
Presentation Control, Scan Rate, Flash Duty
F_DEFAULT
Family Code Selection
P.C
Parameter Selection
Option Code
Presentation Control
SS SS SS
Presentation mode  Flash mode Force mode
0 1 2
Scan Rate Control
SS SS
Dynamic  Fixed
0 1
Flash Duty Cycle
SS SS SS SS
1/2 duty cycle

2/3 duty cycle 3/4 duty cycle 4/5 duty cycle
L680/L780 laser imagers don’t support this function.
0 1 2 3
SmartStand Power Off Timeout
SS SS SS
3 mins  5 mins 10 mins
Only available for FuzzyScan Laser model
0 1 2
Presentation Control: When the scanner is placed on SmartStand, the scanner will be switched from hand-held scanning to hands free scanning automatically. Three
hands scanning modes are available. You are recommended to use flash mode or force mode while under insufficient ambient light.
Scan Rate Control: The scanner will have better motion tolerance when you select Fixed scan rate. Its suitable for application which needs higher motion tolerance on
the move. But this may impact to the reading distance.
The Flash Duty Cycle is designed to control the flashing frequency of the light source.
The SmartStand Power Off Timeout is a pre-defined duration for scanner’s light source on time when the scanner is placed on SmartStand. While the scanner is placed on
SmartStand, the scanning-type will be switched from hand-held scanning to presentation scanning and the light source will be forced on automatically. The light source will be off when the pre-defined duration is up.
Page 62
FuzzyScan
Configure Your FuzzyScan
56
PROGRAM
Operation Control
(L Series)
Laser Imager, LED Illumination Control
F_DEFAULT
Family Code Selection
P.C
Parameter Selection
Option Code
LED Illumination Control
SS SS
Always on Intelligent Mode

Only available for L Series.
0 1
LED Illumination Delay
SS SS SS SS SS
100 ms 150 ms  200 ms 250 ms 300 ms
Only available for L Series.
0 1 2 3 4
LED Illumination Control: When you enable always on, the LED illumination will be always on when you press the trigger. When you enable intelligent mode, the scanner will emit the laser aiming line first, the LED illumination will be turned on after the preset LED illumination delay. Intelligent mode is recommended to be used in regular ambient light environment.
Page 63
FuzzyScan
Configure Your FuzzyScan
57
PROGRAM
Operation Control
(F & L Series)
Laser Aiming Control , 1D Barcode Reading Direction
F_DEFAULT
Family Code Selection
P.C
Parameter Selection
Option Code
Laser Aiming Control
SS SS
Disable Enable 
Only available for L Series.
0
1
1D Barcode Forward-reading
Indication
SS SS
MS
None ◆ S
User defined character(1 character)
0
1
2 〔00-7F〕
1D Barcode Backward-reading
Indication
SS SS
MS
None X
User defined character(1 character)
0
1
2 〔00-7F〕
1D Barcode Direction
Indication Transmission
SS SS SS SS
Disable

Enable prefix direction mark transmission Enable suffix direction mark transmission Enable both prefix and suffix direction mark transmission
0 1 2 3
Laser Aiming Control: You can disable or enable laser aiming line when you scan PDF barcode.
Page 64
FuzzyScan
Configure Your FuzzyScan
58
PROGRAM
Operation Control
(A Series)
Operation , Presentation and Illumination Control
F_DEFAULT
Family Code Selection
P.C
Parameter Selection
Option Code
Operation Mode
SS SS SS SS SS SS SS SS SS
Low Power mode (Low power triggering) Trigger mode (External triggering)

Presentation mode (Auto detection) Alternative mode (Periodic power off) Force mode (Continued power on) Toggle mode (Repeat reading) Diagnostic mode (Test reading) Level mode (Auto power off) Multiple Read Mode
0 1 2 3 4 5 6 7 8
Presentation Control
SS SS
Presentation mode  Force mode 0 1
Illumination Control
SS SS
Disable Enable 
0
1
Presentation Background Lighting
SS SS
LEDs Off LEDs On

0
1
Hand-Held Mode: Low power mode, Trigger mode, Alternative mode, Toggle mode, Level mode, Multiple read mode
Hand-Free Mode: Presentation mode, Force mode,
The Illumination Control is only available for hand-held mode.
Presentation Background Lighting Control: You can enable or disable presentation background lighting of the scanner according to the ambient light condition in presentation mode. When the ambient light is dim or dark, you can enable this function to turn on the scanners LED illumination at a dim level. This is helpful for scanner to detect the motion of scene.
Page 65
FuzzyScan
Configure Your FuzzyScan
59
PROGRAM
Operation Control
(A Series)
Aiming Control, Delay Aiming & Decode Aiming Control
F_DEFAULT
Family Code Selection
P.C
Parameter Selection
Option Code
Aiming Control
SS SS SS
Regular Aiming Intelligent Aiming  Delay Aiming Control
0 1 2
Delay Aiming Time-out Control
SS SS SS SS SS SS SS SS
200 ms 400 ms

800 ms 1 sec
1.5 secs. 2 secs. 3 secs. 4 secs.
0 1 2 3 4 5 6 7
Decode Aiming Control
SS SS
SS SS
Disable in Hand-Held mode Enable in Hand-Held mode

Disable in Hand-Free mode Enable in Hand-Free mode 
0 1 2 3
The Aiming Control is only available for trigger mode. In Intelligent Aiming, the aiming light is turned on when the scanner is lifted. A trigger pull activates decoding process.
After 2 seconds of inactivity, the aiming light will be shut off. Delay Aiming Control allows a delay time for the operator to aim the scanner before the image is taken. During the delay time, the aiming light will be on, but the LED illumination won’t be turned on until the delay time is up.
The Delay Aiming Time-out Control is only available for trigger mode. You can use Delay Aiming Time-out Control to set the delay time.
Page 66
FuzzyScan
Configure Your FuzzyScan
60
PROGRAM
Operation Control
(A Series)
 Center Alignment, Unique Bar Code Reporting
F_DEFAULT
Family Code Selection
P.C
Parameter Selection
Option Code
Center Alignment
SS SS
SS SS
Disable in Hand-Held Mode

Enable in Hand-Held Mode Disable in Hand-Free Mode

Enable in Hand-free Mode
0 1 2 3
Unique Bar Code Reporting
SS SS
Disable

Enable 0 1
Center Alignment: When this function is enabled, the scanner only decodes barcode(s) around aiming line.
Unique Bar Code Reporting: When this function is enabled, the scanner only outputs unique barcode when the trigger is pressed. This function is workable when Multiple
Read mode is selected.
Page 67
FuzzyScan
Configure Your FuzzyScan
61
PROGRAM
Operation Control
(A Series)
Batch Reading 
F_DEFAULT
Family Code Selection
P.C
Parameter Selection
Option Code
Batch Reading
SS
None  Batch Reading rule input ( 01-16 rules)
[FIN]
[Rules] [FIN]
Batch Reading: When this function is enabled, you can scan multiple barcodes one by one continuously upon one trigger event. The scanner reports a good read beep and
indication only if all bar codes set by the Batch Reading Rule” are read. Otherwise, the scanner reports an error beep and indication. The scanned data will be transmitted according to the preset sequence which is defined in “Batch Reading Rule” regardless the scanned order of those barcodes.
- Batch Reading function is only available Trigger Mode.
-
Batch Reading is not available when Multiple Read Mode or Center Alignment is turned on.
-
Batch Reading Rule:
- To set the Batch Reading rule
1. Scan the PROGRAM symbol.
2. Scan the Batch Reading symbol (Family Code).
3. Use the Option Code to define the preset Batch Reading rule.
4. Scan the FIN symbol.
5. Scan the END symbol to save your Batch Reading rule. Note: Scan the ABORT and END symbol to exit without saving any Batch Reading rule setting.
- When you scan “None”, the preset Batch Reading Rule will be cancelled.
- Batch Reading Rule Syntax:
[n] [Element 1] FF [Element 2] FF [Element 3] FF …[Element n] FF Where n is the number of elements in the overall rule. The number of elements is up to 16.FF indicates the end of one element.
- Element structure:
[Cino ID Hex value] [Code length] [Character match(es)]
Where: [Cino ID Hex value]
Length: 2 byte Please find Cino ID hex value from Symbology ID Table in appendix. Locate the Hex value for the symbology and scan the 2 digit hex values from the Option Code.
Note: 99 is the universal number, indicating all symbologies.
Page 68
FuzzyScan
Configure Your FuzzyScan
62
[Code length]
Length: 4 byte Specify what length of data output will be acceptable for this symbology. When you calculate the length, you must consider the whole data string which includes the programmed Preamble, Postamble, Scanned Data Length, Prefix/Suffix Symbol ID or AIM ID. Scan the four digit data length from the Option Code.
Note: 40 characters is entered as 0040; 9999 is a universal number, indicating all lengths.
[Character match]
Length: 2-8 byte You can refer to HEX/ASCII Reference Table to find the Hex value that represents the character(s) you want to match. Use the Option Code to scan the alphanumeric combination that represents the ASCII characters. You can match up to 4 characters which are counted from the start character of the whole Data String. Note: When setting the matched character(s), you must match the content of the whole Data String, including the programmed Preamble, Postamble, Scanned
Data Length, Prefix/Suffi Symbol ID OR AIM ID if you had defined. FF is the universal character, indicating all characters.
- Batch Reading rule example In this example, you are scanning Code 39, Code 128, and Code 93 bar codes, but you would like to output the data in following sequence: Code 128 - Code 39 - Code 93
B-CODE39
A-CODE128
C-CODE93
Page 69
FuzzyScan
Configure Your FuzzyScan
63
You would set up the Batch Reading rule with the following command line:
[PROGRAM] [Batch Reading] [0301999941FF07999942FF09999943FF] [FIN] [END] The breakdown of the command line is shown below:
03
The number of elements in the overall rule
01
Code identifier of Code 128
9999
Code length that must match for Code 128, 9999 = all lengths
41
Start character match for Code 128, 41h = “A”
FF
End of first code
07
Code identifier of Code 39
9999
Code length that must match for Code 39, 9999 = all lengths
42
Start character that must match for Code 39, 42h = B
FF
End of second code
09
Code identifier of Code 93
9999
Code length that must match for Code 93, 9999 = all lengths
43
Start character match for Code 93, 43h = C
FF
End of third code
To program the previous example using specific lengths, you would have to count the programmed Preamble, Postamble, Scanned Data Length, Prefix/Suffix Symbol ID OR AIM ID if you had defined as part of the length. If you enable the Suffix Symbol ID of symbology, you would add one character to the previous example’s length.
You would set up the Batch Reading rule with the following command line:
[PROGRAM] [Batch Reading] [0301001041FF070009FF09000943FF] [FIN] [END]
Page 70
FuzzyScan
Configure Your FuzzyScan
64
The breakdown of the command line is shown below:
03
The number of elements in the overall rule
01
Code identifier of Code 128
0010
Code length that must match for Code 128 A-CODE128 sample length (9) plus Suffix Symbol ID (1) = 10
41
Start character match for Code 128, 41h = “A”
FF
End of first code
07
Code identifier of Code 39
0009
Code length that must match for Code 39 B-CODE39 sample length (8) plus Suffix Symbol ID (1) = 9
FF
Universal matched character, indicating all character Also indicate end of second code
09
Code identifier of Code 93
0009
Code length that must match for Code 93 C-CODE93 sample length (8) plus Suffix Symbol ID (1) = 9
43
Start character match for Code 93, 43h = C
FF
End of third code
Note: If the [Character match(es)] is set to “FF”, the following “FF” which indicated the end of the code was not need to set.
- Structure of Data String
STX
(RS232/USB COM
interface)
Preamble
Scanned Data
Length
Prefix Symbol ID
Or
Prefix AIM Symbol ID
Scanned Data modified
by DataWizard
Suffix Symbol ID
Or
Suffix AIM Symbol ID
Postamble
ETX
(RS232/USB COM
interface)
1 character
1-15 characters
2-4 digits
1 or 3 characters
Variable length
1 or 3 characters
1-15 characters
1 character
Page 71
FuzzyScan
Configure Your FuzzyScan
65
PROGRAM
Operation Control
(All Series)
Buzzer, Indicator & Vibrator Control 
F_DEFAULT
Family Code Selection
P.C
Parameter Selection
Option Code
Buzzer Tone Adjust
SS SS SS SS SS SS SS
Buzzer tone – mute Buzzer tone – low (Frequency 1.20 kHz) Buzzer tone – medium (Frequency 2.70 kHz)

Buzzer tone – high (Frequency 2.81 kHz) Buzzer tone - extremely high (Frequency 2.93 kHz) Power-on beep  No power-on beep
0 1 2 3 4 5 6
Power On Indicator
SS SS SS
Disable (LED off) LED steady on

LED flash
F560 series scanner doesnt support this function.
0 1 2
Good Read Indicator
SS SS
Disable Enable 
0 1
Vibrator Control
SS SS
Disable Enable  Optional function is only available for vibrator model.
0 1
Buzzer Tone Adjust: Available firmware: A680 / A780 1.00.01 and above
A770 rev.1.00.26, rev. 2.00.10 and above A670 rev 1.00.07 and above F680 / L680 rev.2.01.12 and above F780 / L780 rev.2.01.12 and above F560 rev. 2.01.14 and above
Page 72
FuzzyScan
Configure Your FuzzyScan
66
PROGRAM
Operation Control
(A Series)
Buzzer Volume
F_DEFAULT
Family Code Selection
P.C
Parameter Selection
Option Code
Buzzer Volume
SS SS SS
Low Medium High 
0 1 2
Buzzer Volume: Available firmware: A680 / A780 1.00.01 and above
A770 rev. 1.00.21, rev. 2.00.04 and above A670 rev 1.00.03 and above
Page 73
FuzzyScan
Configure Your FuzzyScan
67
PROGRAM
Operation Control
(All Series)
Redundancy & 1D Code Inverse Reading 
F_DEFAULT
Family Code Selection
P.C
Parameter Selection
Option Code
Redundancy
SS SS SS SS SS SS
None Level 1  Level 2 Level 3 Level 4 Level 5
To prevent potential miss reading.
0 1 2 3 4 5
1D Barcode Inverse Reading
SS SS
Disable  Enable
0 1
The Redundancy is the number of times the same bar code label has to be decoded before it is transmitted.
Page 74
FuzzyScan
Configure Your FuzzyScan
68
PROGRAM
Operation Control
(All Series)
Reread Delay & Good Read Delay Control 
F_DEFAULT
Family Code Selection
P.C
Parameter Selection
Option Code
Reread Delay
(Double Scan Verification)
SS SS SS SS SS SS
Disable Immediate time out  Short time out Medium time out Long time out Force verification
0 1 2 3 4 5
Good Read Delay
SS SS SS SS SS SS SS
None  200 msec. 500 msec. 1 sec.
1.5 sec. 2 sec. 3 sec.
0 1 2 3 4 5 6
The Reread Delay (Double Scan Verification) is designed to inhibit FuzzyScan from reading the same bar code label twice in pre-defined short duration. Force Verification
will not allow reading of the same bar code twice.
This Good Read Delay is the minimum amount of time before the imager can read another bar code.
Page 75
FuzzyScan
Configure Your FuzzyScan
69
PROGRAM
Operation Control
(All Series)
Light Source On Time, Hands Free Time-out 
F_DEFAULT
Family Code Selection
P.C
Parameter Selection
Option Code
Light Source On Time
SS SS SS SS
Short  Medium Long Extremely long
0 1 2 3
Hands Free Time-out
SS SS SS SS SS
Short  Medium Long Extremely long Disable
0 1 2 3 4
Good Read Duration
SS SS SS SS SS
Short Medium

Long Extremely long Extremely short
0 1 2 3 4
Time Delay to Low Power Mode
SS SS SS SS SS SS
1 sec 3 secs 5 secs 7 secs 9 secs Immediate 
0 1 2 3 4 5
The Light Source On Time is a pre-defined light source time out counter for Alternative Mode, Presentation Mode and Level Mode. The scanner keeps the light source on till the pre-defined light source on time is up. You can adjust this parameter to meet your own application requirement.
The Presentation Mode, Force Mode and Flash Mode are referred to as “hands free” mode. The hands free mode will be automatically changed to manual trigger mode when you press the trigger. You can remain the scanner in manual trigger mode by setting the Hands Free Time-Out. Once the time-out duration is up (if theres no any trigger operation), the imager will revert to the original hands free mode.
The Time Delay to Low Power Mode sets the time for scanner to enter low power mode after any scanning activity. This setting is only available for the scanner is in low power mode.
Page 76
FuzzyScan
Configure Your FuzzyScan
70
PROGRAM
Operation Control
(All Series)
Presentation Auto-Sense & Sensitivity
F_DEFAULT
Family Code Selection
P.C
Parameter Selection
Option Code
Presentation Auto-sense
SS SS
Disable Enable  0 1
Presentation Sensitivity
SS SS SS SS SS SS SS
Level 1 Level 2 Level 3 Level 4 Level 5

Level 6 Level 7
0 1 2 3 4 5 6
When enabling the Presentation Auto-sense, the scanner can be switched from hand-held and hand-free scanning automatically when working with the SmartStand.
The presentation sensitivity is used to configure the sensitivity level when the scanner is set as presentation mode. The higher lever means higher sensitivity for detecting
the barcode.
Page 77
FuzzyScan
Configure Your FuzzyScan
71
PROGRAM
Condensed DataWizard
Preamble, Postamble, Data Length & Symbol ID Trans. 
F_DEFAULT
Family Code Selection
P.C
Parameter Selection
Option Code
Preamble
SS
MS
None

1-15 characters
Maximum 15-character input; scan “FIN” to terminate this selection.
FIN
[00-7F], [FIN]
Postamble
SS
MS
None

1-15 characters
Maximum 15-character input; scan “FIN” to terminate this selection.
FIN
[00-7F], [FIN]
Data Length Transmission
SS SS
Disable

Enable 2~4 digits data length transmission
0 1
Symbology ID Transmission
SS SS SS SS SS SS SS
Disable symbology ID transmission

Enable prefix symbology ID transmission Enable suffix symbology ID transmission Enable both prefix and suffix symbology ID transmission Enable prefix AIM symbology ID transmission Enable suffix AIM symbology ID transmission Enable both prefix and suffix AIM symbology ID transmission
0 1 2 3 4 5 6
DataWizard is the most powerful, Artificial-Intelligence based data editing expert system provided specially for the FuzzyScan family bar code readers. Through DataWizard, you can process the scanned data prior the transmissions in many ways as: Insert, Delete, Match, Verify, Replace, Reorganize, and Repeat Transmission. It will help you to arrange the transmission of scanned data to any specific format without software modification.
Due to the resources used by this system, Full-feature DataWizard is only supported by PowerTool. Through the PowerTool, all settings and configurations can be done on-screen on Windows 2000 / XP / 7
A Condensed Version DataWizard is provided by each FuzzyScan series. Through this menu, the condensed DataWizard can be utilized via bar code menu readings with ease.
Please note that all “Character” input should be referred to the ASCII/HEX Table listed in Appendix to find matched HEX value.
If you have any problem to use DataWizard, please refer to following pages for details and consult your local FuzzyScan vendor or our web site for any assistance.
Page 78
FuzzyScan
Configure Your FuzzyScan
72
PROGRAM
Condensed DataWizard
Data Formatter Setting 
F_DEFAULT
Family Code Selection
P.C
Parameter Selection
Option Code
2nd Option Code
Formatter Control
SS
MS MS
Disable

Select one bar code symbology Select all bar code symbologies
FIN
(2 digits)
00
automatic termination automatic termination
1st Insertion
SS
DS
Disable

Enable
2-digits identified position; max. 3 insertion characters
FIN
(2 digits)
position
[1-3 characters], [FIN]
2nd Insertion
SS
DS
Disable

Enable
2-digits identified position; max. 3 insertion characters
FIN
(2 digits)
position
[1-3 characters], [FIN]
3rd Insertion
SS
DS
Disable

Enable
2-digits identified position; max. 3 insertion characters
FIN
(2 digits)
position
[1-3 characters], [FIN]
4th Insertion
SS
DS
Disable

Enable
2-digits identified position; max. 3 insertion characters
FIN
(2 digits)
position
[1-3 characters], [FIN]
The Data Formatter is used to edit the scanned raw data prior to transmitting the data to the host computers or terminals. It allows you to select desired bar code symbologies for formatter control, and provides Multiple Position Insertion and Multiple Character Insertion (max three characters) in the identified position.
While the Data Formatter is enabled, it arranges only scanned data without Preamble, Postamble, STX, ETX, Data Length, Prefix/Suffix Symbology ID or Record Suffix. All of the above programmable parameters perform the same function depending on your setting.
Regarding the “Bar Code Selection” and “Position Calculation” of data formatter, please refer to page 75 for details.
Please note that all “Character” input should be referred to the ASCII/HEX Table listed in Appendix to find matched HEX value.
Page 79
FuzzyScan
Configure Your FuzzyScan
73
PROGRAM
Condensed DataWizard
Data Verifier Setting 
F_DEFAULT
Family Code Selection
P.C
Parameter Selection
Option Code
2nd Option Code
Verifier Control
SS
MS MS
Disable

Select one bar code symbology Select all bar code symbologies
FIN
(2 digits)
00
automatic termination automatic termination
Identified Data Length
SS
DS
Disable

Enable
Determine the identified data length for verification.
FIN
(2 digits)
1st Identified Character
SS
DS
Disable

Enable
2-digits checking position; 1 identified character
FIN
(2 digits)
position
[00-7F]
2nd Identified Character
SS
DS
Disable

Enable
2-digits checking position; 1 identified character
FIN
(2 digits)
position
[00-7F]
3rd Identified Character
SS
DS
Disable

Enable
2-digits checking position; 1 identified character
FIN
(2 digits)
position
[00-7F]
The Data Verifier is used to provide advanced verification for error-free scanning and to work as an Embedded Data Transmitting Filter.
All data must conform to the Identified Bar Code Symbologies, Identified Data Length, and one to three Identified Characters in the checking position. Otherwise, the FuzzyScan will not transmit the data to the host computers or terminals, but will instead issue 3 long beeps for verification error and skip the scanned data.
The Data Verifier checks only scanned data without Preamble, Postamble, STX, ETX, Data Length, Prefix/Suffix Symbology ID or Record Suffix.
Regarding the “Bar Code Selection” and “Position Calculation” of Data Verifier, please refer to page 75 for details.
Please note that all “Character” input should be referred to the ASCII/HEX Table listed in Appendix to find matched HEX value.
Page 80
FuzzyScan
Configure Your FuzzyScan
74
PROGRAM
Condensed DataWizard
Data Replacer Setting 
F_DEFAULT
Family Code Selection
P.C
Parameter Selection
Option Code
2nd Option Code
Replacer Control
SS
MS MS
Disable

Select one bar code symbology Select all bar code symbologies
FIN
(2 digits)
00
automatic termination automatic termination
1st Replacement
SS
DS
Disable

Enable
2-digits identified position; 1 replacement character
FIN
(2 digits)
position
[00-7F]
2nd Replacement
SS
DS
Disable

Enable
2-digits identified position; 1 replacement character
FIN
(2 digits)
position
[00-7F]
3rd Replacement
SS
DS
Disable

Enable
2-digits identified position; 1 replacement character
FIN
(2 digits)
position
[00-7F]
The Data Replacer is used to edit the scanned raw data prior to transmitting the data to the host computers or terminals. It allows you to select desired bar code symbologies for replacer control, and provides Multiple Position Replacement in the identified position.
All data must conform to the Identified Bar Code Symbologies, and one to three Identified Characters in the identified position. While the Data Replacer is enabled, it arranges only scanned data without Preamble, Postamble, STX, ETX, Data Length, Prefix/Suffix Symbology ID or Record Suffix.
Regarding the “Bar Code Selection” and “Position Calculation” of Data Replacer, please refer to page 75 for details.
Please note that all “Character” input should be referred to the ASCII/HEX Table listed in Appendix to find matched HEX value.
Page 81
FuzzyScan
Configure Your FuzzyScan
75
PROGRAM
Condensed DataWizard
Data Organizer Setting 
F_DEFAULT
Family Code Selection
P.C
Parameter Selection
Option Code
2nd Option Code
Organizer Control
SS
MS MS
Disable

Select one bar code symbology Select all bar code symbologies
FIN
(2 digits)
00
Automatic termination Automatic termination
1st Organization
SS
DS
Disable

Enable
2-digits identified position; Forward/backward data transmission setting
FIN
(2 digits)
position direction
0 (Forward)

1 (Backward)
2nd Organization
SS
DS
Disable

Enable 2-digits identified position; Forward/backward data transmission setting
FIN
(2 digits)
position direction
0 (Forward)

1 (Backward)
Include/Exclude Control
SS
DS
Transmitted data excluded the data of identified position

Transmitted data included the data of identified position
0 1
The Data Organizer is used to edit the scanned raw data prior to transmitting the data to the host computers or terminals. It allows you to select desired bar code symbologies for organizer control, and provides maximum two identified positions to send the data forward or backward. It also allows you to control the transmitted data including or excluding the data of identification position. Please refer to the application example listed in page 75 for details.
While the Data Organizer is enabled, it arranges only scanned data without Preamble, Postamble, STX, ETX, Data Length, Prefix/Suffix Symbology ID or Record Suffix.
Regarding the “Bar Code Selection” and “Position Calculation” of Data Organizer, please refer to page 75 for details.
Please note that all “Character” input should be referred to the ASCII/HEX Table.
Page 82
FuzzyScan
Configure Your FuzzyScan
76
Select a Bar Code Symbology
You can select one or all types of bar code symbologies to use Condensed DataWizard for advanced transmission
arrangement. If you scan “00” to select all types, the
FuzzyScan will arrange all incoming data to meet your pre-defined format. If you want to select only one type bar code, please select one of the option code listed below.
1D Bar Code Symbology
Code 128
01
Matrix 2 of 5
38
GS1-128
31
Interleaved 2 of 5
48
UPC-A
02
China Postal Code
58
UPC-A with 2 supplement
32
German Postal Code
68
UPC-A with 5 supplement
42
Standard/Industrial 2 of 5
08
UPC-E
03
Code 93
09
UPC-E with 2 supplement
33
Code 11
10
UPC-E with 5 supplement
43
MSI/Plessey
11
EAN-13
04
UK/Plessey
12
EAN-13 with 2 supplement
34
Telepen
13
EAN-13 with 5 supplement
44
GS1 DataBar
14
EAN-8
05
GS1 DataBar Limited
22
EAN-8 with 2 supplement
35
GS1 DataBar Expanded
23
EAN-8 with 5 supplement
45
Composite Codes
24
Codabar/NW-7
06
IATA
15
Code 39
07
Coupon Code
16
Code 32
37
PDF417
17
Trioptic Code 39
47
Micro PDF417
25
Codablock F
18
Code 16K
19
Code 49
20
2D Bar Code Symbology
QR Code
A0
MaxiCode
A2
MicroQR Code
A0
Aztec Code
A3
DataMatrix
A1
Chinese Sensible Code
A4
GS1 DataMatrix
A5
Postal Code
Korea Post Code
21
Japanese Post
B4
Australian Post
B0
KIX Post
B5
British Post
B1
Planet Code
B6
Intelligent Mail barcode
B3
Postnet
B8
Position Calculation
[Data Formatter]
If there is a 5-character input data string, refer to the following to calculate the actual position for insertion:
X X X X X
00 01 02 03 04 05
[Data Verifier, Data Replacer, Data Organizer]
If there is a 11-character data string, please refer to the following to calculate the actual position for identification.
X X X X X X X X X X X
00
01
02
03
04
05
06
07
08
09
10
Page 83
FuzzyScan
Configure Your FuzzyScan
77
Application Example
If your bar code label is a 16-digit Interleaved 2 of 5 which includes the information of 6-digit date code, 6-digit serial number and 4-digit unit price, you want the FuzzyScan do the following for you without software modification:
Apply only Interleaved 2 of 5 to the condensed
DataWizard.
 Check bar code is actually with 16-digit length.  Allow bar code output whose date code is leading with “9”. Three outputs with “TAB” suffix.  The date code output should skip “9” and replaced it by
“A”.
The serial number output should be led with “SN”.  The unit price output should be skipped the first 2 digits.  Test Bar Code : 9 8 1 0 2 5 1 2 3 4 5 6 9 8 7 6
Actual Output : A81025[TAB]SN123456[TAB]76[TAB]
Programming Procedure
[Data Verifier]
Scan “Program” to enter the programming mode. Scan “Verifier Control” and set bar code symbology to
48” (Interleaved 2 of 5).
Scan “Identified Data Length” and set the length to
16”.
Scan “1st Identified Character” and set the identified
position to “00”, then set the identified character to “39
(Hex Code of 9).
[Data Formatter]
Scan “Formatter Control” and set bar code symbology
to “48”.
Scan 1st Insertion” and set the identified position to
06”, then inserted characters to “09” (Hex Code of TAB), “53” (Hex Code of S), “4E” (Hex Code of N).
Scan “2nd Insertion” and set the identified position to
12”, then inserted character to “09”. In the final, you must scan “FIN” (Finish) code to terminate this
selection.
Scan “3rd Insertion” and set the identified position to
16”, then inserted character to “09”. In the final, you must scan “FIN” (Finish) code to terminate this
selection.
Page 84
FuzzyScan
Configure Your FuzzyScan
78
[Data Replacer]
Scan “Replacer Control” and set bar code symbology to
48”.
Scan “1st Replacement” and set the identified position to
00”, then replaced character to “41” (Hex Code of A).
[Data Organizer]
Scan “Organizer Control” and set bar code symbology to
48”.
Scan “1st Organization” and set the identified position to
16”, then set the data transmission to “0” (forward).
Scan “2nd Organization” and set the identified position to
17”, then set the data transmission to “1” (backward).
Scan “END” (Exit) to terminate the programming.
[Important Notice]
Please note that Condensed DataWizard will follow the preset working flow as below:
Verifier Formatter Replacer Organizer
So when you set the identified position in Data Organizer, you must consider the inserted data which you already set via Data Formatter.
Page 85
FuzzyScan
Appendix
79
Symbology ID Table
Each AIM Code Identifier contains the three-character string ]cm where: ] = Flag Character; c = Code Character; m = Modifier Character
1D Symbology ID Table
Code Family
Primary Format
Cino ID
AIM ID
Code Family
Primary Format
Cino ID
AIM ID
Hex
Value
Char.
Code
Char.
Modified
Char.
Hex
Value
Char.
Code
Char.
Modified
Char.
UPC
UPC-A
2
A
E 0 EAN/JAN
EAN/JAN-8
05
N
E
4
UPC-A with 2 supple.
32 1 EAN/JAN-8 with 2 supple.
35 1 UPC-A with 5 supple.
42 2 EAN/JAN-8 with 5 supple.
45
2
UPC-E
3 E 0
EAN/JAN-13
04
F
E
0
UPC-E with 2 supple.
33 1 EAN/JAN-13 with 2 supple.
34
1
UPC-E with 5 supple.
43 2 EAN/JAN-13 with 5 supple.
44
2
Example: A UPC-A bar code 012345678950 with 2 supplement 12 is transmitted as ]E0012345678950]E112
Example: A EAN/JAN-8 bar code 49123562 with 5 supplement 12345 is transmitted as ]E449123562]E212345
Code 128
Code 128
01
B C m
Code 93
Code 93
09 H G m GS1-128
31 C 1
Code 11
Code 11
10 P H
m
Codabar
Codabar/NW-7
06 D F m MSI/Plessey
MSI/Plessey
11 R M
m
Code 25
Standard/Industrial 2 of 5
08 I S 0 UK/Plessey
UK/Plessey
12 S P
0
Matrix 2 of 5
38 K X 0 Telepen
Telepen
13 T B
m
Interleaved 2 of 5
48 J I m GS1 DataBar
GS1 Databar
14
X e m
GS1 DataBar Limited
22
GS1 DataBar Expanded
23
China Postal Code
58 L X
0
Composite
Composite Code
24
Code 39
Code 39
07 G A m German Postal Code
68 M I m Code 39 Trioptic
47 W X
0
IATA
IATA
15 O R m Code 32
37 G A
0
UCC Coupon
UCC Coupon Code
Z
PDF417
PDF417
17
V L m
Micro PDF417
25
Example : A UPC-A 512345678900 + GS1-128 81010123451297 bar code is transmitted as]E0512345678900]C181010123451297
Example: A EAN-13 9923456789019 + GS1-128 81010123451297 bar code is transmitted as ]E09923456789019]C181010123451297
Codablock
Codablock F
18 Y O
m
Korea Post
Korea Post Code
21 a X
0
Remark: Above examples are given for the transmission of AIM ID.
Page 86
FuzzyScan
Appendix
80
2D Symbology ID Table
Code Family
Primary Format
Cino ID
AIM ID
Code Family
Primary Format
Cino ID
AIM ID
Hex
Value
Char.
Code
Char.
Modified
Char.
Hex
Value
Char.
Code
Char.
Modified
Char.
QR Code
QR Code
A0 b Q
m
British Post
British Post
B1
h X 0
Micro QR Code
Micro QR Code
Intelligent Mail barcode
Intelligent Mail barcode
B3 j 0
Data Matrix
Data Matrix
A1
c d m
Japanese Post
Japanese Post
B4 k 0
GS1 Data Matrix
A5
MaxiCode
MaxiCode
A2 d U m KIX Post
KIX Post
B5 l 0
Aztec Code
Aztec Code
A3 e z m Planet Code
Planet Code
B6 m 0
Chinese Sensible
Chinese Sensible
A4
f X 0
Postnet
Postnet
B8 o 0
Australian Post
Australian Post
B0 g 0
Page 87
FuzzyScan
Appendix
81
Keyboard Function Code Table
No.
ANSI
ASCII
Key Function
Ctrl Output
No.
ANSI
ASCII
Key Function
Ctrl Output
00
NUL
00H
RESERVED
Ctrl + @
16
DLE
10H
F7
Ctrl + P
01
SOH
01H
CTRL (Left)
Ctrl + A
17
DC1
11H
F8
Ctrl + Q
02
STX
02H
ALT (Left)
Ctrl + B
18
DC2
12H
F9
Ctrl + R
03
ETX
03H
SHIFT
Ctrl + C
19
DC3
13H
F10
Ctrl + S
04
EOT
04H
CAPS LOCK
Ctrl + D
20
DC4
14H
F11
Ctrl + T
05
ENQ
05H
NUM LOCK
Ctrl + E
21
NAK
15H
F12
Ctrl + U
06
ACK
06H
ESC
Ctrl + F
22
SYN
16H
INS (Insert) (Edit)
Ctrl + V
07
BEL
07H
F1
Ctrl + G
23
ETB
17H
DEL (Delete) (Edit)
Ctrl + W
08
BS
08H
BACK SPACE
Ctrl + H
24
CAN
18H
HOME (Edit)
Ctrl + X
09
HT
09H
TAB
Ctrl + I
25
EM
19H
END (Edit)
Ctrl + Y
10
LF
0AH
F2
Ctrl + J
26
SUB
1AH
PAGE UP (Edit)
Ctrl + Z
11
VT
0BH
F3
Ctrl + K
27
ESC
1BH
PAGE DOWN (Edit)
Ctrl + [
12
FF
0CH
F4
Ctrl + L
28
FS
1CH
UP (Edit)
Ctrl + \
13
CR
0DH
ENTER (CR)
Ctrl + M
29
GS
1DH
DOWN (Edit)
Ctrl + ]
14
SO
0EH
F5
Ctrl + N
30
RS
1EH
LEFT (Edit)
Ctrl + 6
15
SI
0FH
F6
Ctrl + O
31
US
1FH
RIGHT (Edit)
*see note
To emulate the keyboard function key input for user definable parameters, user must configure actual content using the Reserved ASCII 00 31 characters, and also Enable the “Function Key Emulation”. Otherwise, the Ctrl output will be done by the scanner. Please refer to the above Keyboard Function Code Table which is for IBM PC/XT/AT, PS/2, PS/VP, COMPAQ PC, HP Vectra PC, Notebook PC, APPLE and PowerMac, and WYSE PC Enhanced or fully compatible machines.
The last character in the Ctrl Output column is varied for different countries.
Country (refer to Keyboard Layout) & Character
United State
-
Switzerland
-
France
=
Belgium
-
UK - Germany
-
Sweden
-
Denmark
-
Norway
-
Spain
-
Italy -
Page 88
FuzzyScan
Appendix
82
ASCII Input Shortcut
To configure the user definable parameters of FuzzyScan via programming menu, FuzzyScan will ask you to scan your desired ASCII value in HEX form. You have to refer to the
HEX/ASCII Table” for details.
Example:
If you want the scanned data output leading with a Dollar
Sign, you have to set the “Preamble” to “$”. The configuration
procedure is listed below for reference.  Scan the system command – PROGRAM listed on page
3-24 to enter programming mode.
Scan family code – PREAMBLE to select this family. Refer to the Hex/ASCII Table, you will find the HEX value
of “$” is 24.
Scan the option code – 2 listed on the fold out back cover. Scan the option code – 4 listed on the fold out back cover. Scan the system command – FIN (Finish) to terminate
Preamble setting.
 Scan the system command – End to exit the programming
mode for normal operation.
HEX/ASCII Reference Table
H
L
0 1 2 3 4 5 6
7
0
NUL
DLE
SPACE 0 @ P `
p 1 SOH
DC1 ! 1 A Q a q 2 STX
DC2 " 2 B R b r 3 ETX
DC3 # 3 C S c s
4
EOT
DC4 $ 4 D T d t
5
ENQ
NAK % 5 E U e u 6 ACK
SYN & 6 F V f v 7 BEL
ETB ' 7 G W g w 8 BS
CAN ( 8 H X h x 9 HT
EM ) 9 I Y i y A LF
SUB * : J Z j z B VT
ESC + ; K [ k { C FF
FS , < L \ l | D CR
GS - = M ] m } E SO
RS . > N ^ n ~ F SI
US / ? O _ o DEL
Example : ASCII “A” HEX “41”; ASCII “a” “61”
: High Byte of HEX Value : Low Byte of HEX Value
Page 89
FuzzyScan
Appendix
83
Host Interface Quick Set
RS232 Serial
PS/2 (DOS/V) KBW Standard Mode
USB HID Standard Mode
Keyboard Replacement
PS/2 (DOS/V) KBW Turbo Mode
USB HID Turbo Mode
USB HID Legacy Mode
USB Com Port Emulation
Page 90
FuzzyScan
Appendix
84
Operation Mode Quick Set (F & L Series)
Low Power (Low power trigger)
Force (Continued power on)
Alternative (Periodic power off)
Trigger (External triggering)
Toggle (Repeat reading)
Flash (Pulse driven reading)
Presentation (Auto sensing)
Diagnostic (Test reading)
Level (Auto power off)
Page 91
FuzzyScan
Appendix
85
Operation Mode Quick Set (A Series)
Low Power (Low power trigger)
Force (Continued power on)
Alternative (Periodic power off)
Trigger (External triggering)
Toggle (Repeat reading)
Level (Auto power off)
Presentation (Auto sensing)
Diagnostic (Test reading)
Multiple Read
Page 92
FuzzyScan
Appendix
86
Option Codes
0 A
5
F
1 B
6
2 C
7
3 D
8
4 E
9
FIN (Finish)
Abort
END (Exit)
Page 93
FuzzyScan
Appendix
87
System Commands
PROGRAM
(Enter Programming Mode)
System Information List
(SYSLIST)
Save User Default
FIN (Finish)
PowerTool Host Link
User Default
END
(Exit Programming Mode)
Factory Default
Master Default
Factory Default: After scanning Factory Default command, all parameters will be returned to factory default value.
Master Default: After scanning “Master Default” command, the scanner will remain the pre-set parameters of Host Interface Selection, Keyboard Interface Control (except Record Suffix; Preamble; Postamble), Serial Interface Control (except Record Suffix; Preamble; Postamble), and Wand/Laser Emulation Control, the rest of parameters will be returned to default value.
User Default: After scanning Save User Default command, all current parameters will be stored to the flash memory. Once you change the parameter and would like to return to previous setting, please scan User Default.
Page 94
FuzzyScan
Appendix
88
Loading...