Chauvin Arnoux F07 User Manual

Page 1
■■
Pince multimètre
■■
■■
Clamp multimeter
■■
■■
Vielfachmesszange
■■
■■
Pinza multimetro
■■
■■
Pinza multimetrica
■■
F 07
Notice de fonctionnement User's manual Bedienungsanleitung Libretto d’Istruzioni Manual de Instrucciones
1
Page 2
English
Meaning of symbol Caution! Please consult the User Manual before using the device.
In this User Manual, failure to follow or carry out instructions preceded by this symbol may result in personal injury or damage to the device and the installations.
Meaning of symbol This appliance is protected by double insulation or reinforced insulation. It does not ha ve to be connected to an earth protection terminal for electrical safety .
Meaning of CA T III symbol
This voltage surge category III clamp, with pollution level 2, complies with stringent reliability and availability requirements, corresponding to fixed industrial and domestic installations (see IEC 664-1, ed 92).
Thank you for purchasing this “F07” series multimeter c lamp. To get the best service from this instrument:
n read this user’s man ual carefully, n respect the safety precautions detailed
PRECAUTIONS FOR USE
Never use on networks with voltage higher than 600 V in relation to the ground and whose overvoltage category is higher than III, i.e.
fixed industrial and domestic installations (see IEC 664-1).
Indoor use in environments with a degree of pollution of 2 maximum (cf. IEC 664-1), temperature of 0°C to + 50°C and relative humidity
lower than 70%.
Use accessories compliant with safety standards (NF EN 61010-2-031) with minimal voltage
category III.
Never open the clamp bo x bef ore disconnecting all power sources .
Never connect to the circuit to be measured if the clamp box is not
properly closed.
Before any measurement, chec k and switch.
When measuring current check for proper alignment of the conductor in relation to the markers and proper closing of the jaws.
Always disconnect the clamp from any po wer source before changing the battery.
Do not perform resistance tests, continuity tests or semi-conductor tests on a circuit under power .
the proper positioning of the cables
of 600 V and ov ervoltage
25
Page 3
CONTENTS
1. PRESENTATION .................................................................... 27
2. DESCRIPTION ....................................................................... 27
3. IMPLEMENT ATION - FUNCTIONAL CHARA CTERISTICS ... 31
3.1 Reference conditions...................................................... 31
3.2 V oltage measurements
3.3 Audio continuity test (
3.4 Resistance measurement (Ω)......................................... 33
3.5 Semi-Conductor T est (
3.6 Current measurements (A) ......................................... 34
3.7 INRUSH function ............................................................ 35
3.8 Frequency measurements (Hz) ...................................... 35
3.9 Temperature Measurement (T°) ..................................... 36
3.9.1 No sensor.......................................................... 36
3.9.2 With sensor ....................................................... 36
3.10 Adapter Function (Adp.) ................................................. 37
3.11 Secondary Functions ..................................................... 39
3.11.1 Display lock ....................................................... 39
3.11.2 Preselection of MIN/MAX mode ........................ 39
3.11.3 Automatic compensation for cable resistance ... 39
3.11.4 Automatic compensation
of current measurement zero ............................ 39
3.11.5 Manual mode selection AC, DC or AC + DC ..... 39
3.11.6 Selections possible in continuity fonction .......... 39
3.11.7 Selection OF INRUSH function ......................... 40
3.11.8 Choice of units (°C or °F)
in measurement of temperature ........................ 40
3.11.9 Memorisation of unit (°C or °F)
in measurement of temperature ........................ 40
3.11.10 Suppression of automatic stop .......................... 40
3.11.11 Activation of the V-Live function ......................... 40
3.11.12 Modification of audio indication
threshold in continuity test ................................. 40
3.11.13 Programming the scale factor
of adapter function............................................. 41
3.11.14 Default configuration of the unit ......................... 41
3.11.15 Date of last calibration carried out on unit.......... 41
3.11.16 Display of the internal software version ............. 41
3.11.17 Displaying of display segments.......................... 41
4. GENERAL SPECIFICA TIONS................................................ 42
4.1 Dimensions and weight .................................................. 42
4.2 Clamp Tightening Capacity ............................................ 42
4.3 Pow er supply.................................................................. 42
4.4 Environmental parameters ............................................. 42
4.5 Compliance with norms .................................................. 42
4.6 V ariations in operating range .......................................... 43
4.7 Marginal operating conditions......................................... 45
5. TO ORDER ............................................................................. 46
6. WARRANTY ........................................................................... 46
7. MAINTENANCE ..................................................................... 47
7.1 Changing the battery ...................................................... 47
7.2 Storage........................................................................... 47
7.3 Cleaning ......................................................................... 47
7.4 Metrological verification .................................................. 47
8. APPENDIX ........................................................................... 117
(V) ......................................... 31
) .............................................. 32
) ............................................ 33
26
Page 4
1. PRESENTATION
Clamp multimeter F07 emphasises reliability and simplicity of use to respond to the needs of power professionals:
A compact unit integrating the current sensor for intensity
measurements without breaking the test circuit
Outstanding ergonomic features, in particular:
- automatic or manual selection of the type of signal to be measured, AC or DC,
- measurement of the RMS value of any signal (AC + DC)
- automatic selection of measurement calibre,
- programmable audio voltage indication: V-Live,
- «measurement range exceeded» indication,
- Backlighting of the digital display
- auto power off system,
- MIN - MAX – PEAK value recording function,
- correction of deviations in DC measurement (DC zero)
- automatic compensation of measurement cable resistance ( zero).
Compliance with CEI electrical safety standards and CE
markings
Lightness and ruggedness for field use
With all new additional features:
"Inrush" function in particular for measurement of motor
starting currents.
"Adapter" function with direct reading via scale factor
programming.
2. DESCRIPTION
(see diagram in § 8 Appendix)
Jaws
➀ ➁
6- w ay rot a r y sw it ch :
OFF Deactivation of the clamp , activation is ensured by
selection of other functions
V Measurement of DC and AC voltages (rms value)
Continuity measurement, by use of the yellow resistance key and semi-conductor test
A Measurement of DC and AC voltages (rms value) Measurement of clamp ambient or external
Adp. Selection of «Adapter» function
temperature, depending on presence or absence of sensor, in °C or °F.
27
Page 5
Command keys
The keys are capable of 3 types of action:
Short pressure
< 1.3 s, it is valid if the key pressure is detected.
Long pressure
> 1.3 s, this gives access to a measurement or operating mode. Holding or releasing the key has no effect.
Held pressure
Gives access to a measurement or operating mode and remains in this mode as long as pressure is held. Releasing the key causes return to the previous mode.
■■
HOLD has 4 different functions
■■
(see description § 3.11)
:
- Display lock
- Preselection of MIN/MAX mode
- Automatic compensation for cable resistance
- Automatic compensation of current measurement z ero
■■
The yellow key has 4 different functions
■■
(see description § 3.11)
:
- M anual mode selection AC, DC or AC + DC
- Selection of resistance function ( ), semi-conductor test ( ), continuity test ( )
- Selection OF INRUSH function
- Choice of units (°C or °F) in temperature measurement
■■
MIN/MAX operates by end-around shift on short
■■
pressure:
MIN/MAX V, A and Adp. Functions Other functions
1st press PEAK value MAX value 2nd press MAX value MIN value 3rd press MIN value Return to MAX value
4th press Return to MAX value _
At any time, a long press on the ke y will quit the MIN/MAX mode.
Note: In MIN/MAX mode, the Automatic stop function of the unit is unavailable (symbol P)
■■
Hz A short press displays the frequency of the
■■
measured signal, another press switches back to the previous value . This key is active only for the A and V functions in AC or AC + DC mode.
■■
Short pressure: display backlight command.
■■
Automatic shutdown after 2 minutes. Held pressure: display of estimated remaining
battery power, e xpressed in hours (except INR USH and phase order functions).
28
Page 6
HOLD key / switch combination
(see description § 3.11):
- Suppression of automatic stop function
- Activation of the V-Live function
- Display of the internal software version
Yellow key / switch combination
(see description § 3.11):
- Modification of audio indication threshold in continuity test
- Choice of default units (°C or °F) in temperature measurement
-
Programming of scale factor while in Adapter function
- D efault configuration of the unit
MIN/MAX key / switch combination
(see description § 3.11):
- Date of last calibration performed on unit
Liquid crystal display
The liquid crystal display includes the digital display of the measured values, the related units and symbols.
Digital display
4 digits, 9999 points, 3 decimal points , + and - signs (DC and peak measurement).
+OL
: Positive value range exceedance (> 3999
points)
-OL OL
- - - -
: Negative value range exceedance : Unsigned value range exceedance : Indeterminate value (middle segments)
Symbol display
+
-
P
Flashing, clamp power limited to approximately 1 hour Steady, battery drained, clamp operation or
accuracy no longer guaranteed
Constant operation (no automatic shutdown)
: ON steady when the INRUSH function is
selected
Fixed: Continuity measurement Flashing: V-Live function selected
29
Page 7
HOLD
HOLD Function active
PEAK ON in V, A and Adp. in MIN/MAX mode if the
measurement of the peak value is selected
MAX Indicates the display of a maximum value in
MIN/MAX mode
MIN Indicates the display of a minimum value in
MIN/MAX mode
AC Fixed: measurement in AC manual mode
Flashing: measurement in A C automatic mode
DC Fixed: measurement in DC manual mode
Flashing: measurement in DC automatic mode
AC + DC measurement in AC + DC manual mode Temperature measurement INT Temperature measurement when the
terminals are not connected or if the connected thermocouple is defective
EXT Temperature measurement when the
thermocouple is connected Semi-conductor test on position
m Scale factor < 1 in adapter function k Scale factor > 1000 in adapter function
■■
The Buzzer
■■
Different tones are emitted according to the function given to the buzzer :
- Shor t and medium sound: valid key
- Shor t and high-pitched sound: prohibited key
- Shor t and low sound : Quit MIN/MAX mode
- 2 short and high-pitched beeps: validation of a
configuration parameter
- Short and medium sound every 400 ms: voltage measured higher than the unit’s guaranteed safety voltage .
- 5 short and medium recurring beeps: automatic deactivation of the instrument
- Shor t and medium sound: measured continuity value, lower than programmed threshold, short-circuit junction during semi-conductor test.
- Modulated medium continuous sound: value measured in volts, higher than 45 V peak when the V-Live function is selected.
30
Page 8
3. IMPLEMENTATION
FUNCTIONAL CHARACTERISTICS
3.1 Reference conditions
The functional characteristics mentioned in each of the measurement functions are guaranteed in the following reference ranges:
- Temperature: +23°C ± 3 K
- Humidity: 45% to 75% relative humidity
- Supply voltage: 8.5 V ± 0.5 V
- Frequency range of applied AC signal: 45 -65 Hz
- Peak factor of applied AC signal: √ 2
- Position of conductor in clamp jaws: centred.
- Conductor diameter: ≤ 5 mm
- No external AC magnetic field
- No electrical field
3.2 Voltag e measurements (V)
1. Connect the measurement leads to the instrument’s
terminals, complying with the polarities indicated: red lead on the «+» terminal and black lead on the "COM" terminal.
2. Set the rotary switch to position "V ".
3. Connect the assembly to the voltage source to be
measured, making sure if possible that this voltage does not exceed the maximum acceptable limits (see table below). Range switching and AC/DC selection are automatic. Actuate the yellow ke y to force A C/DC or AC + DC selection manually if necessary.
If the signal measured is > 45 V peak, the audio indication is activated if the V-Live function is selected
Display range 40 V 400 V 4.000 V
Measuring range
Accuracy 1% L +5 pt 1% L +2 pt 1% L +2 pt Resolution 10 mV 0.1 V 1 V Input impedance 1 M
Protection 600 V AC or DC
(2)
0.2 V to 40.0 V to 400 to 600 V
39.99 V 399.9 V 400 to 900 Vpeak
(see § 3.11.11)
(1)
(1) In DC, the display indicates +OL above +600 V and -OL
above -600 V (900 V in PEAK mode). In AC and AC + DC , the display indicates OL ov er 600 Vrms (900 V in PEAK mode).
(2) In AC if the v alue of the voltage measured is < 0.15 V the
display indicates 0.00.
.
For voltages the buzzer indicates that the measured voltage is
higher than the unit’s guaranteed safety voltage.
≥≥
600 VDC or rms, a repetitive beep of
≥≥
31
Page 9
■■
MIN/ MAX Mode:
■■
- Accuracy: same as previous table +0.2% L
- Capture time: 100 ms typ.
■■
PEAK Mode:
■■
- Accuracy: same as previous table +2% L
- Capture time: 500 µs typ. (2.5 ms max.)
■■
Special characteristics in V-Live mode
■■
- Detection threshold accuracy: 45 Vpeak ± 2V
3.3 Audio continuity test ( )
1. Connect the measurement cables to the unit terminals.
2. Set the rotary switch to position " "
3. Connect the unit to the circuit to be tested. The buzzer is
continuously active as soon as contact is established (circuit closed) and if the resistance value measured is lower than the threshold value chosen by the progr amming (adjustable from 1 to 40 , see § 3.11.12). Above 400 Ω, the display indicates
OL
.
■■
Cable resistance compensation (
■■
ΩΩ
zero)
ΩΩ
To measure low resistance values, measure the cable resistance first.
- Short-circuit the cables.
- Press and hold the HOLD key until zero appears on the display. The cable resistance value will then be memorised and subtracted from the value of the resistance measured later.
Note: If the value measured is higher than 2 Ω, this
correction is stopped and the memorised correction value is reset to zero.
■■
Characteristics
■■
Display range 400
Measuring range 0.0 to 399.9
(1)
Accuracy Resolution 0.1 Open circuit voltage 3.2 V Measuring current 320 µA
Protection 500 V A C or 750 V (DC or peak)
: 1% L +2 pt
(1) with compensation for measurement cable resistance
■■
MIN/ MAX Mode:
■■
- Accuracy: same as previous table +0.2% L
- Capture time: 100 ms typ.
32
Page 10
3.4 Resistance measurement (
Ω)
1. Connect the measurement cables to the unit terminals.
2. Set the rotary switch to position " " and press once
on the yellow key: The symbol is no longer displayed.
3. Connect the unit to the resistance to be tested. Range selection is automatic. To measure low resistance with accuracy compensate the measurement cable resistance
(see § 3.3).
Above 40 , the display indicates OL.
Display range 400 4000 Ω 40 k
Measuring range 0.0 to 400 to 4.00 kΩ à
399.9 3999 W 39.99 k Accuracy Resolution 0,1 1 10 Open circuit voltage 3.2 V Measuring current 320 µA 40 µA
Protection 500 V A C or 750 V (DC or peak)
(1)
: 1% L +2 pt
(1) With compensation for measurement cable resistance
■■
MIN/ MAX Mode:
■■
- Accuracy: same as previous table +0.2% L
- Capture time: 100 ms typ.
3.5 Semi-Conductor Test ( )
1. Connect the measurement leads to the instrument’s
terminals, complying with the polarities indicated: red lead on the «+» terminal and black lead on the «COM» terminal.
2. Set the rotary switch to position and press twice on the yellow key: The symbol appears on the display.
3. Connect the unit to the semi-conductor (junction) to be tested. The measurement current moves from the «+» terminals to the «COM» terminal. It corresponds to the direct testing of the semi-conductor junction.
- Shor t-circuit junction: audio indication for a threshold
< 0.050 V
- Cut or reversed junction (or threshold > 3.2 V) : display
of OL. Display range 4 V
Measuring range 0 to 3.199 V Accuracy 1% L +2 pt Resolution 1 mV Measuring current Protection 500 V AC or 750 V (DC or peak)
(1)
2 mA to 4 mA
(1) Per the voltage measured
■■
MIN/ MAX Mode:
■■
- Accuracy: same as previous table +0.2% L
- Capture time: 100 ms typ.
33
Page 11
3.6 Current measurements (A)
1. Set the rotary switch to position "A ".
2. Clamp the conductor carrying the current to be measured,
check for proper closing of the jaws and for foreign bodies in the gap. For DC , the "" arrow engraved on the ja ws must be directed in the presumed direction of current circulation for the sign of the displayed value to be significant. Range switching and AC/DC selection are automatic. Actuate the yellow ke y to force A C/DC or AC + DC selection manually if necessary.
■■
Correction of the current zero (DC zero)
■■
To measure the low currents, perform a zero correction first.
- Press and hold the HOLD key until zero appears on the
display. The corrected value will then be memorised and subtracted from the value of the current measured later.
Note: this correction is performed only on the DC component of the zero. If the value measured is higher than 6 Α, this correction is stopped and the memorised correction value is reset to zero.
■■
Characteristics
■■
Display range 40 A 400 A 4000 A
Measuring range
Accuracy
Resolution 10 mA 100 mA 1 A
(3)
(2)
0.20 to 40.0 to 400 to
39.99 A 399.9 A 600 A
1.5% L +10 pt
1.5% L +2 pt
(1)
peak
(1)In DC, the displa y indicates +OL above +400 A and -OL
above -400 A (600 A in PEAK mode). In AC and AC + DC, the display indicates OL over
400 Arms (600 A in PEAK mode).
(2)In AC, if the value of the current measured in AC is <
0.15 A, the display shows 0.00.
(3)With correction of zero in DC
- Repeatability of the measurement after se veral consecutive closings of the clamp: 0.3% typical
■■
MIN/ MAX Mode:
■■
- Accuracy: same as previous table +0.2% L
- Capture time: 100 ms typ.
■■
PEAK Mode:
■■
- Accuracy: same as previous table +2% L +0.5 A
- Capture time: 500 µs typ. (2.5 ms max.)
34
Page 12
3.7 INRUSH function
■■
Description
■■
This function is used to follow quick changes in the current, such as a damped sinusoidal quantity, by measuring the successive rms values calculated on ½, 1, 2½, 5 and 10 periods from the largest rms value computed and updated on ½ period. The applications are:
- Measurement of motor star t-up currents
- Correct definition of fuses and circuit breakers (signal
amplitude-time relationship)
- Stress on components by current overload The field of application is limited to industrial frequencies (15 Hz... 70 Hz)
■■
Implementation
■■
This function is accessible in AC or AC + DC current measurement, after selection of the MIN/MAX mode.
Action Display Comments
Press 0.5 P then the value the key for rms Enter the function yellow corresponding
out F
Press on 1P-2, 5 P - 5P - 10P-0,5P HOLD key Consultation of values then press with each time rms (computed on successively the rms value of consecutive periods ) the yellow corresponding key alternately
Signal frequency < 15 Hz or > 70 Hz
Short pressure Return to values Exit from the function on the MIN, MAX or PEAK return to MIN/MAX mode MIN/MAX key
■■
Characteristics
■■
- Accuracy: 5% +0.5 A
- Capture time: 10 periods of the signal frequency (200 ms
at 50 Hz)
- Range for use: 5 A peak f or the first period of the signal
3.8 Frequency measurements (Hz)
This function is active only for the A and V functions in AC (AC or AC + DC ) mode.
1. Perform a short pressing on the Hz key, the display shows the frequency of the measured signal.
2. Pressing again returns to the previously displayed measurement.
35
Page 13
■■
Characteristics
■■
Display range 40 Hz 400 Hz 4000 Hz 40 kHz Measuring range
Accuracy 0.4% L +1 pt Resolution 0.01 Hz 0.1 Hz 1 Hz 10 Hz
T riggering threshold
(1)
10.00 to 40.0 to 400 to 4 .00 kH z to
39.99 Hz 399.9 Hz 3999 Hz 19.99 kHz
(2)
5 V or 10 A
(1) Below 5 Hz, the display shows 0.0 (2) Below the triggering threshold, the display shows an
indeterminate value (- - - -). In AC + DC mode ,the triggering threshold corresponds to the AC component (AC) of the signal.
■■
MIN/ MAX Mode:
■■
- Accuracy: same as in the table above +0.2% of reading
with limitation to 5 kHz.
- Capture time: 125 ms typ. every 400 ms
3.9 Temperature Measurement (T°)
3.9.1 No sensor
Set the rotary switch to position "T°". The temperature displayed is the internal temperature of the unit (INT symbol ON), equivalent to the ambient temperature after a sufficient period of thermal stabilisation. It can be expressed in °C or °F: choice of a unit using the yellow key.
3.9.2 With sensor
1. Connect sensor (K couple) to the clamp terminals, observing the polarity indicated and placing it where you wish to measure the temperature.
2. Set the rotary switch to position "T°". The temperature displayed is that of the sensor (EXT symbol ON), it may be expressed in °C or °F: choice of unit by the yellow key.
■■
Characteristics
■■
Function Internal temperature External temperature Type of sensor Integrated Circuit K Couple Display range 400°C 400°C 4000°C
Extended +50,0°C +399,9°C +1.000°C measurement +15,0°F to -50,0°F to +400°F to
Accuracy ± 1,5°C 1% L ±1,5°C 1% L ±1,5°C
Resolution 0,1°C1°C
Detection of INT symbol ON sensor shutdown - instead of EXT
Thermal time 0.7 min./°C Depending on sensor model constant
400°F 400°F 4000°F
-10,0°C to -50,0°C to +400°C to
+120,0°F +399,9°F +1832°F
±2,7°F 1% L ±2,7°F 1% L ±2,7°F
0,2°F1°F
Note: The accuracy announced for external temperature measurement does not integrate the accuracy of the K couple.
36
Page 14
■■
MIN/ MAX Mode:
■■
- Accuracy: same as previous table +0.2% L
- Capture time: 100 ms typ. (every 800 ms)
3.10 Adapter Function (Adp.)
This function is used to connect any adapter ensuring the conversion of a physical quantity (mechanical, electrical, etc...) into a DC or AC current 4 V.
This function also has a programmable scale factor by decade from 0.001 to 100,000 capable of direct reading (with possible display of the multiple: "m" for "milli" and "k" for "kilo"), for the adapters having a transfer ratio equal to a whole multiple of 10.
■■
Scale factor selection
■■
The table below shows the diff erent sensitivities of a direct­reading adapter after selection of a scale factor.
Sensitivity S (mV/A) Scale factor
(example in Amperes) to be programmed 10 mV/kA (0.01 mV/A) 10 k 100 mV/kA (0.1 mV/A) 100 k
1 mV/A 1
10 mV/A 10
100 mV/A 100
1000 mV/A (1 mV/mA) 1 m
10 mV/mA 10 m
100 mV/mA 100 m
The example given in Amperes (A) is valid for any other quantity: humidity (% HR), light (lux), speed (m/s), etc...
■■
Scale factor programming
■■
1. Maintain the yellow key pressed and select the "Adp." position on the rotary switch: the display indicates the programmed scale factor (1 by default) as follows:
Display 1 m 10 m 100 m 1 10 100 10 k 100 k
2. To change the scale factor, press the yellow key successively until the required value is obtained.
3. Use the rotary switch to memorise this value.
Implementation
1. Connect the adapter to the clamp
2. Program the scale factor if necessary (see § above)
3. Set the rotary switch to position "Adp".
4. Select the type of voltage using the yellow key if
necessary (AC or DC)
37
Page 15
■■
Characteristics
■■
Display range
(1)
4000 Measuring range Accuracy 1% L +2 pt Input impedance 1 M
Protection 600 V A C or DC
(1)
The basic display is 4000 points. The position of the
(2)
0.0 to 399.9 mV 0.400 to 3.999 V
decimal and the display of multiples (m and k) depend on the programming of the scale factor.
(2)
- In DC, the display shows +OL above +3999 points and -OL above -3999 points.
- In AC, the display shows OL above +3999 points.
- Range switching (400 mV to 4 V) is automatic.
- In AC, the lower limit of the measurement range is
5.0 mV.
■■
MIN/ MAX Mode:
■■
- Accuracy: same as previous table +0.2% L
- Capture time: 100 ms typ.
■■
PEAK Mode:
■■
- Accuracy: same as previous tab le +3% of reading in DC
(3% L +20 mV in AC)
- Capture time: 500 µs typ. (2.5 ms max.)
38
Page 16
3.11 Secondary Functions
3.11.1 Display lock
A short pressing on the HOLD key freezes the display. A second pressing unlocks the display.
3.11.2 Preselection of MIN/MAX mode
A short pressing on the HOLD key then on the MIN/MAX key preselects the MIN/MAX mode. A second pressing on the HOLD key makes the MIN/MAX mode effective.
This function is used to select the MIN/MAX mode upon request to prevent, for example, the unwanted or mistaken integration of MIN/MAX values.
3.11.3 Automatic compensation for cable resistance
This compensation is obtained by holding down the HOLD key when the continuity test or measurement resistance function is selected.
When the key is released, when the display shows zero, the correction value is memorised in the clamp.
If the value measured is higher than 2
stopped and the memorised correction value is reset to zero . This correction is prohibited in MIN/MAX mode.
ΩΩ
Ω, this correction is
ΩΩ
3.11.4 Automatic compensation of current measurement zero
This compensation is obtained by holding down the HOLD key when the current measurement function is selected.
When the key is released, when the display shows zero, the correction value is memorised in the clamp.
If the value measured is higher than 6 A, this correction is stopped and the memorised correction value is reset to zero . This correction is prohibited in MIN/MAX mode.
3.11.5 Manual mode selection AC, DC or AC + DC
By default the clamp s witches to AC or DC mode automatically (AC or DC symbol flashing) for the V and A functions.
Short successive pressures on the yellow key are used to select manually AC measurement, DC measurement, measurement of compound signals (AC + DC) and return to automatic mode. When the mode is selected manually, the symbol AC, DC or AC + DC is fixed. The selection of this manual mode is impossible in MIN/MAX or HOLD mode. The selection of AC + DC mode for the adapter function is not possible.
3.11.6 Selections possible in continuity fonction
By default the clamp is in continuity function ( ) for the position corresponding to the rotary switch. Successive presses on the yellow key enable selection of the resistance measurement (), then the semi conductor
test function ( ) and to return to the continuity function ( ).
39
Page 17
3.11.7 Selection OF INRUSH function
This is done in function A (AC, AC+DC) by pressing first on the MIN/MAX key then on the yellow key. Consultation of the values corresponding to this function is possible by pressing first on the HOLD key then by short successive presses on the yellow key
(see § 3.7)
. Quitting this function is possible by short presses on the MIN/ MAX key.
3.11.8 Choice of units (°C or °F) in measurement
of temperature
This is done when the temperature function is selected by short presses on the yellow key, enabling the selection of °C or °F as applicable. This unit is not memorised when the clamp is stopped. See § 3.11.9 if you wish to memorise this unit.
3.11.9 Memorisation of unit (°C or °F) in measurement
of temperature
Hold the yellow key pressed and bring the rotary switch from the OFF position to the position.
The unit emits a double beep then the T° symbol goes ON and the °F symbol flashes if the unit was previously in °C or the °C symbol flashes if it was in °F.
The selected configuration is memorised when the key is released: the °F or °C symbol is then on steady.
3.11.10 Suppression of automatic stop
Hold the HOLD key pressed and bring the rotary switch from the OFF position to the position .
The unit emits a double beep then the
symbol flashes.
P
The selected configuration is memorised when the key is released: the P symbol remains lit continuously.
The unit returns to automatic stop mode when the switch goes to the OFF position.
3.11.11 Activation of the V-Live function
Hold the HOLD key pressed and bring the rotary switch from the OFF position to the V position.
The unit emits a double beep then the V and symbol flashes. The selected configuration is memorised when the key is
released: the V symbol is then ON steady and the symbol flashes.
Proceed in the same way to suppress the V-Live function: the
symbol then goes OFF when the key is released.
3.11.12 Modification of audio indication threshold in
continuity test
Hold the Yellow key pressed and bring the rotary switch from the OFF position to the position .
The unit beeps, the Ω and symbols go ON along with the threshold value (40.0 by default)
.
40
Page 18
Adjustment is then possible from 1 to 40 by successive pressings on the yellow key (short pressure: prog ression of 1 by 1 ; press and hold: progression of 10 by 10 Ω). Once the value is chosen, actuate the rotary switch to memorise.
3.11.13Programming the scale factor of adapter function
Hold the Y ellow key pressed and bring the rotary switch from the OFF position to the Adp position.
The unit emits a beep and the scale factor value is displayed (1 per default).
Adjustment of the scale factor is then possible by successive pressings on the yellow key of 1 m (0.001) to 100 k (100,000)
(see § 3.10)
. Once the value is chosen, actuate the rotary switch to memorise.
3.11.14 Default configuration of the unit
Hold the Y ellow key pressed and bring the rotary switch from the OFF position to the A position.
The unit emits a double beep then all the segments of the digital display and the symbol flash.
The default configuration is memorised when the key is released: the display no longer flashes and the
symbol disappears. The default configuration is:
- Audio identification threshold: 40
- Auto cut-off with
- V-Live function: none
- Scale factor in adapter function: 1
- Unit for temperature measurement: not controlled
3.11.15 Date of last calibration carried out on unit
Hold the MIN/MAX key pressed and bring the rotary switch from the OFF position to the V position.
The unit emits a beep, then the calibration date of the unit is displayed in the form "week - year" (SS.AA) as long as the MIN/MAX key is pressed.
3.11.16 Display of the internal software version
Hold the HOLD key pressed and bring the rotary switch from the OFF position to the A position.
The unit beeps, the software version is displayed in the form UX.XX for 2 seconds, then replaced by the displaying of all the segments of the display as long as the HOLD key is pressed.
3.11.17 Displaying of display segments
See § 3.11.16. It is usually possible for any key-switch combination not described previously.
41
Page 19
4. GENERAL SPECIFICATIONS
4.1 Dimensions and weight
70 x 193 x 37 mm 260 g
4.2 Clamp Tightening Capacity
26 mm
4.3 Power supply
A standard 9 V alkaline battery (type CEI 6LF22, 6LR61
or NEDA 1604). To change it, see §7.1.
Average battery charge life: 75 h or 25,000 x 10 s.
measurement
Battery charge indicator
Flashing: charge life < 1 h Fixed: Change battery
Automatic OFF after 10 minutes with no action on the s witch
or the keys (reactivation by switching from the OFF position of the switch or by actuating any key)
4.4 Environmental parameters
■■
Temperature - Humidity
■■
+
-
:
90 80 70
Relative humidity in % RH
60 50 40 30 20 10
0
-50
-40 -30 -20 -10 0 10 20 30 40 50 60 70 80 90
Temperature in °C
Reference range Operating range
Storage range (without battery)
■■
Altitude
■■
- Operation: 2000 m
- Storage 12000 m
■■
Indoor use
■■
Impermeability: protection index IP 40
(according to EN
60529, ed. 92)
4.5 Compliance with norms
■■
Electrical safety
■■
(as per EN 61010-1 ed. 95 and 61010-2-032, ed. 93)
- Dual insulation:
- Installation category: III
- Pollution lev el: 2
- Rated voltage: 600 V (RMS or DC)
o r - Installation category : IV
- Rated voltage : 300V (RMS or DC)
42
Page 20
■■
Electric shocks
■■
(test as per IEC 1000-4-5)
- 6 kV in RCD mode on the voltmeter function, aptitude criterion B
- 2 kV induced on the current measurement cable, aptitude criterion B
■■
Electromagnetic compatibility
■■
(as per EN 61326-1 ed. 97 + A1)
Emission: class B Immunity:
- Electrostatic discharges: 4 kV on contact, aptitude criterion B 8 kV in the air, aptitude criterion B
- Radiated field: 10 V/m, aptitude criterion B
- Fast Transients: 1 kV, aptitude criter ion B
- Conduit interference: 3 V/m, aptitude criterion A
■■
Mechanical resistance
■■
- Free fall 1 m
- Impacts: 0.5 J
- Vibrations: 0.75 mm
■■
Auto power OFF
■■
(test as per IEC 68-2-32)
(test as per IEC 68-2-27)
(test as per IEC 68-2-6)
(per UL94)
- Housing: V0
-Jaws: V0
- Display window: V2
4.6 Variations in operating range
Influence Meas. range QuantityInfluence
quantitiesquantities influenced Typical Maxi Battery voltage 7.5 to 10 VAll< 1 pt 0.2% L +1 pt Temperature 0...50°CV - Adp. 0.05% L/10°C 0.2% L /10°C +2 pt
A 0.3% L/10°C 0.5% L /10°C +2 pt
- 0.5% L /10°C +1.5°C
Hz 0.03% L/10°C 0.1% L /10°C +2 pt
Relative 10...90% HR V - Adp. < 1 pt 0.1% L +1 pt humidity A 0.2% L 0.3% L +2 pt
0.3% L 0.5% L +1.5°C
Hz 0.05% L 0.1% L +2 pt
Frequency 10 Hz...1 kHz
1 kHz...5 kHz 6% L +1 pt 10 Hz...400 Hz 1% L +1 pt
400 Hz...1 kHz A See curve 5% L +1 pt 1 kHz...5 kHz -3 dB
10 Hz...400 Hz 400 Hz...1 kHz 5% L +1 pt
Position ofPosition conductor on in the perimeter A 0.7% L 1% L +1 pt jawsinternal (f ≤ 400 Hz) jaws
Retentivity 0...600 peak A 2 mA/A 3 mA/A
V
Adp
0.1% L/10°C 0.2% L /10°C +2 pt
0.2% L 0.3% L +2 pt
See curve
See curve
1% L +1 pt
1% L +1 pt
43
Page 21
Influence Meas. range Q uantity Influence
quantitiesquantities influenced Typical Maxi
Adjacent Conductor conductor in contact with crossed by external A 45 dB 40 dB a current perimeter
DC or RMS jaws
400 A
Conductor held by the clamp
Application of a voltage on the clamp
Peak 1 to 3.5 limited factor to 600 A peak
Rejection of 0...600 V/50 Hz V
0...400 A DC V - Adp. or RMS T °
0...600 V DC or RMS
900 V peak
A< 1 pt 1 pt
AC, AC + DC) 1% L 3% L +1 pt
A( V(
AC, AC + DC) 1% L 3% L +1 pt
DC 50 dB 45 dB
< 1 pt 1 pt
serial mode 0...4 V/50 H z Adp DC 60 dB 50 dB in DC 0...400 A/50 Hz A DC 40 dB 35 dB
Rejection of 0...600 V DC V(
AC, AC + DC)> 60 dB 50 dB
serial mode 0...4 V D C Adp AC 60 dB 50 dB in AC 0...400 A DC A(AC, AC + DC)> 50 dB 40 dB
Rejection of common mode
0...600 V/50 Hz A 0.07 A/100 V 0.1 A/100 V
V< 1 pt 60 dB
T°< 1 pt 60 dB
Influence of external 0...400 A/m magnetic (50 Hz)
A 70 dB 60 dB
field Number of
manœuvres opening of
50000 A 0.3% L 1% +1 pt
jaws
44
Page 22
■■
T ypical frequency response curve
■■
- V = f (f)
8.00%
6.00%
4.00%
2.00%
0.00%
-2.00%
-4.00%
Error measurement (%)
-6.00%
-8.00%
-10.00% 10 Hz
Specified limits
100 Hz
Frequency (Hz)
- I = f (f)
10.00%
5.00%
0.00%
-5.00%
-10.00%
-15.00%
-20.00%
Error measurement (%)
-25.00%
-30.00%
-35.00% 10 Hz
Specified limits
100 Hz
Frequency (Hz)
1000 Hz
1000 Hz
10000 Hz
10000 Hz
- V (adp) = f (f)
10.00%
5.00%
0.00%
-5.00%
-10.00%
Error measurement (%)
-15.00%
-20.00%
-25.00% 10 Hz
4.7 Marginal operating conditions
■■
Clamped conductor temperature: 110°C
■■
Specified limits
100 Hz
Frequency (Hz)
45
1000 Hz
10000 Hz
Page 23
5. T O ORDER
F07 .................................................................. P01.1209.07Z
Delivered in a blister with set of 2 probe cables, 1 K thermocouple adapter, 1 x 9 V battery, one carrying case and this operating manual
Accessories and spare parts
Set of 2 cables
with probe tips (NF EN 61010)...................... P01. 2950.84
Set of 2 cables
with safety connector (NF EN 61010)........... P01.2950.88
Set of 2 alligator clips (NF EN 61010) .......... P01.1018.48
Set of 2 cables with IP2X touch prods.......... P01.2951.57
Carrying case N° 7 ........................................ P01.2985.32
K couple adapter / 4 mm connector dia. ....... P01.1017.80
General purpose temperature sensor
with SK13-type handle .................................. P03.6529.18
Many other types of K couple sensor are available
depending on the target application (ambient air, surface, penetration...) Please consult us
Different measurement accessories widen the scope or
provide new functions to your clamp... Please consult us
NB: Always use accessories adapted to the voltage and
overvoltage category of the circuit to be measured (per NF EN 61010).
6. WARRANTY
Our guarantee is applicable for twelve months after the date on which the equipment is made available (extract from our General Conditions of Sale, available on request).
46
Page 24
7. MAINTENANCE
For maintenance, use only specified spare parts. The manufacturer will not be held responsible for any accident occuring following a repair done other than by its After Sales Service or approved repairers.
7.1 Changing the battery
The clamp must be disconnected from any external electrical source and must not be in contact with the cable.
Set the switch to OFF
Slide a screwdriver into the slot at the top of the battery
cover (rear of the clamp) and push the battery cover upwards.
Replace the used battery by a 9 V battery, type LF22; observe the polarities.
Install the battery in its housing, then install the battery cover.
7.2 Storage
If the clamp is not used for a period of more than 60 days, remove the battery and store it separately.
7.3 Cleaning
The clamp must be disconnected from any external electrical source and must not be holding the cable.
Clean the unit and jaws with a cloth slightly moistened with soapy w ater. Clean off with a damp cloth. Then dry quickly with a cloth or pulsed air.
Do not splash water onto the clamp.
Keep the gap between the jaws in a state of perfect
cleanliness.
7.4 Metrological verification
It is essential that all measuring instruments are regularly calibrated.
For checking and calibration of y our instrument, please contact our accredited laboratories (list on request) or the Chauvin Arnoux subsidiar y or Agent in your country.
Maintenance
Repairs under or out of guarantee: please return the product to your distributor.
47