INSTRUCTION MANUAL
FieldLogger
with 128k Memory
0141 0315-57 01. April 2009
B+B Thermo-Technik Gmb H | Hei nrich-Hertz-Str. 4 | D-78166 Donaueschingen
Technical changes reserved Fon +49 771 83160 | Fax +49 771 831650 | info@bb-sensors.com | bb-sensors.com 1 / 24

INSTRUCTION MANUAL
FieldLogger
with 128k Memory
0141 0315-57 01. April 2009
B+B Thermo-Technik Gmb H | Hei nrich-Hertz-Str. 4 | D-78166 Donaueschingen
Technical changes reserved Fon +49 771 83160 | Fax +49 771 831650 | info@bb-sensors.com | bb-sensors.com 2 / 24
Table of Contents
1. FOREWORD ................................................................................................................................................... 3
2. GENERAL INFORMATION ............................................................................................................................ 3
3. SYMBOLS EMPLOYED ................................................................................................................................. 3
4. SAFETY INSTRUCTIONS .............................................................................................................................. 4
5. INTENDED USE ............................................................................................................................................. 4
6. DISPOSAL ...................................................................................................................................................... 5
7. SCOPE OF DELIVERY .................................................................................................................................. 5
8. DEVICE DESCRIPT IO N ................................................................................................................................. 5
9. INSTALLATION .............................................................................................................................................. 6
10. OPERATION ................................................................................................................................................. 15
11. TECHNICAL DATA ...................................................................................................................................... 23
12. GENERAL QUESTIONS .............................................................................................................................. 24

INSTRUCTION MANUAL
FieldLogger
with 128k Memory
0141 0315-57 01. April 2009
B+B Thermo-Technik Gmb H | Hei nrich-Hertz-Str. 4 | D-78166 Donaueschingen
Technical changes reserved Fon +49 771 83160 | Fax +49 771 831650 | info@bb-sensors.com | bb-sensors.com 3 / 24
1. FOREWORD
Dear customer,
We thank you for having purchased the FieldLogger Register and are very glad that you decided for a product of B+B Thermo-
Technik GmbH. We hope this product will fully satisfy you and will assist you effectively in your work.
This device has been developed to be technically highly up-to date. This product has been designed in accordance with the regnant
European and German national directives and rules. For a proper and effective usage of the product the customer shall observe the
following Operating Instructions. In the case that against one’s expectations any trouble occurs which you can not resolve yourself,
please contact our service centers or our authorized dealer. We will provide you rapid and competent help to minimize the risk of
long time outfalls.
The following operating instruction is an indispensable part of this Product. It contains important advices for the starting up and
further use of the device.
2. GENERAL INFORMATION
This Operation Manual is intended to serve as an aid in the proper setup, installation and operating of the FieldLogger Register.
All essential details of the equipment and all actions required on the part are clearly presented and explained. We thus ask that you
read this manual carefully before proceeding to work with the equipment. Keep this manual available for ready reference in a
convenient and conspicuous location near the equipment.
3. SYMBOLS EMPLOYED
It is necessary to read the following advices before using the product.
The used symbols in the manual acts first of all as eye catcher for
security risks. The symbols do not replace the security advices. The
text must be read completely.
This symbol designates important advices and tips which are
necessary for the success of a procedure. They have to be followed in
order to get good results.

INSTRUCTION MANUAL
FieldLogger
with 128k Memory
0141 0315-57 01. April 2009
B+B Thermo-Technik Gmb H | Hei nrich-Hertz-Str. 4 | D-78166 Donaueschingen
Technical changes reserved Fon +49 771 83160 | Fax +49 771 831650 | info@bb-sensors.com | bb-sensors.com 4 / 24
This symbol advises the user
of danger for persons,
material or environment. The
text gives information that
must be necessarily followed
Caution against electromagnetic fields
(BGV A8, GUV-V A8/W12)
Caution against hot surfaces
(BGV A8, GUV-V A8/W26)
and hot liquids or substances
Caution against severe cold
(BGV A8, GUV-V A8/W17)
Caution against liquids and
hot substances
Caution against dangerous high electrical
voltage (BGV A8, GUV-V A8/W08)
Caution against dangerous
explosive substances (BGV
A8, GUV-V A8/W02)
Caution against dangerous explosive
atmosphere
(BGV A8, GUV-V A8/W21)
Caution against moving parts
Electronic waste
4. SAFETY INSTRUCTIONS
B+B Thermo-Technik GmbH assume no liability for damages occurred through failure to observe these security advices. A usage
none conform to the instructions given in this manual can damage the device
The use of the measurement system must be restricted to qualified personal.
The system is not adequate for use in atmosphere with explosion risk.
Do not use the system in a high electric or magnetic field area.
Test this equipment prior to each use.
The system must be operated only within the limits given in the technical Data
Exposing the system to hot temperatures (higher than the operating temperature) will cause damages in the electronic circuits and
also damage the housing.
5. INTENDED USE
The use of the unit in fields other than those indicated under “SAFETY INSTRUCTIONS” is not allowed for safety reasons.
This instruction manual does not at all substitute any additional instruction manual of connected accessories.

INSTRUCTION MANUAL
FieldLogger
with 128k Memory
0141 0315-57 01. April 2009
B+B Thermo-Technik Gmb H | Hei nrich-Hertz-Str. 4 | D-78166 Donaueschingen
Technical changes reserved Fon +49 771 83160 | Fax +49 771 831650 | info@bb-sensors.com | bb-sensors.com 5 / 24
6. DISPOSAL
This unit has been marked in accordance with the European Device 2002/96/EC on waste electrical and electronic equipment
(WEEE).
At the end of its useful operating life, dispose of the unit as electrical scrap.
Please ask either B+B Thermo-Technik GmbH or your specialist dealer for information on your local collection point. Within the
scope of application if this Directive, B+B Thermo-Technik GmbH is responsible for proper disposal of this unit.
7. SCOPE OF DELIVERY
1 x FieldLogger Register
1 x Operating Manual
1 x FieldLogger Software
8. DEVICE DESCRIPT IO N
FieldLogger is an equipment that reads and registers analogue variables. It works as an electronic data logger, storing logged
information in its internal memory for later analysis. It can also work as a real-time remote measuring module, which does not have
internal memory and only logs information and transmits them to a supervisory system. Its 8 configurable input channels, digital input
and alarm outputs, backup power, etc, make it very versatile, which makes possible to use it in a large number of applications. It is
configured through an easy-to-use software. The equipment is provided with the software, which runs on Windows98 platform and
higher.

INSTRUCTION MANUAL
FieldLogger
with 128k Memory
0141 0315-57 01. April 2009
B+B Thermo-Technik Gmb H | Hei nrich-Hertz-Str. 4 | D-78166 Donaueschingen
Technical changes reserved Fon +49 771 83160 | Fax +49 771 831650 | info@bb-sensors.com | bb-sensors.com 6 / 24
9. INSTALLATION
9.1 INPUT
The logger has 8 input channels. You can configure the type of signal to be read by each channel. All types of input channels
accepted are factory calibrated and no further adjustment is required. The user defines and configures the channels that will be in
use, and disables the others.
The types of input channels accepted by the logger and their maximum ranges are listed in Table 01.
-50.0 to 760.0°C (-58.0 to 1400.0°F)
-90.0 to 1370.0°C (-130.0 to 2498.0°F)
-100.0 to 400.0°C (-148.0 to 752.0°F)
-35.0 to 720.0°C (-31.0 to 1328.0°F)
-90.0 to 1300.0°C (-130.0 to 2372.0°F)
0.0 to 1760.0°C (-32.0 to 3200.0°F)
0.0 to 1760.0°C (-32.0 to 3200.0°F)
150.0 to 1820.0°C (302.0 to 3308.0°F)
-200.0 to 530.0°C (-328.0 to 986.0°F)
Thermocouple J curve, -50.0 to 760.0°C
Thermocouple K curve, -90.0 to 1370.0°C
Thermocouple T curve, -100.0 to 400.0°C
Thermocouple E curve, -35.0 to 720.0°C
Thermocouple N curve, -90.0 to 1300.0°C
Thermocouple R curve, 0.0 to 1760.0°C
Thermocouple S curve, 0.0 to 1760.0°C
Thermocouple B curve, 150.0 to 1820.0°C
Pt100 curve, -200.0 to 530.0°C
Programmable indication range from -1999 to 9999
Programmable indication range from -1999 to 9999
MEASUREMENT INTERVAL
The speed of input channels readings depends on the number of channels enabled and type of input programmed for each channel.
Each channel requires a reading time of about 50 ms. Programmed channels with Pt100 input need m ore than 50 ms for cable
length compensation calculations. 150 ms are required for internal processing, regardless of the number of channels enabled. Thus,
if only one channel is enabled, readings will take place at every 200 ms (50 + 150). If this channel is programmed to Pt100, 250 ms
are required between readings (50 + 50 + 150). Required time to read the eight channels is 550 ms. If they are configured as Pt100
input, required time will be 950 ms.
9.2 OUTPUT
The logger has two relay outputs (ALM1 and ALM2) that can work as Alarm or Digital Output.
When configured as Alarm, they act as configured by the user.
As Digital Outputs they are remotely controlled through serial communication.

INSTRUCTION MANUAL
FieldLogger
with 128k Memory
0141 0315-57 01. April 2009
B+B Thermo-Technik Gmb H | Hei nrich-Hertz-Str. 4 | D-78166 Donaueschingen
Technical changes reserved Fon +49 771 83160 | Fax +49 771 831650 | info@bb-sensors.com | bb-sensors.com 7 / 24
9.3 ALARMS
The logger has two alarms (Alarm 1 and Alarm 2) that work independently. Each alarm has a setpoint that defines the triggering
point. They can act as Low alarm or High alarm.
Low alarms are triggered (turn on an output) when the value read by the input channel is below the setpoint value. The High alarm
is triggered when the value measured is above the setpoint value.
Alarm outputs can be associated to ALM1 (relay 1) and ALM2 (relay 2). The same alarm can be associated to both outputs. Such
settings must be defined for each of the eight input channels.
9.4 SETUP
It is necessary to configure the logger before starting operation. It is necessary to define the working mode assigning values to a
series of parameters. The register is provided with the Configurator software. It provides the parameters that define how the logger
will work.
The software is supplied on CD The installation on the user’s PC is performed through the FL_Setup.exe application.
MINIMUM REQUIREMENTS
To obtain a good performance of the Configurator software, the PC must support the following requirements:
• Pentium III (or equivalent) or higher.
• Windows 98® or higher.
• A free serial port to support the communication with the register.
The system will perform better in computers with higher specifications. A RS485 (serial or USB) converter will be required.

INSTRUCTION MANUAL
FieldLogger
with 128k Memory
0141 0315-57 01. April 2009
B+B Thermo-Technik Gmb H | Hei nrich-Hertz-Str. 4 | D-78166 Donaueschingen
Technical changes reserved Fon +49 771 83160 | Fax +49 771 831650 | info@bb-sensors.com | bb-sensors.com 8 / 24
RUNNING THE LOGGER
Before running the Configurator the logger must be connected to the computer. Use the free serial port, a RS232 / RS485 or USB /
RS485 converter and suitable cables for the connection.
Figure 1 – Overview of the communication system
Note: The logger basic configuration is factory set (communication address 1, baud rate 9600, no parity, etc). The first
configuration must be accomplished in a point-to-point connection with the computer. After it is configured with a specific
communication address, among other parameters, it can be installed in a communication network together with other devices.
When you try to run the Configurator software there is an attempt of immediate communication with the connected logger. If
communication is successful, the Configurator is fully loaded and the logger configuration is read and displayed. Settings are then
allowed to be changed.
In case there is a communication fault, only the communication screen is loaded. It displays the communication parameters used in
communication attempts. Selected serial port, baud rate, address and parity.
For a new communication attempt, the user must:
• Check connections;
• Check if the serial port displayed in the communication screen is the same as the selected one;
• Check if the serial port selected is really free in the computer;
After checking those items, press Search in the Automatic Search (valid only for a point-to-point connection). If communication is
successful, the Configurator software is fully loaded.

INSTRUCTION MANUAL
FieldLogger
with 128k Memory
0141 0315-57 01. April 2009
B+B Thermo-Technik Gmb H | Hei nrich-Hertz-Str. 4 | D-78166 Donaueschingen
Technical changes reserved Fon +49 771 83160 | Fax +49 771 831650 | info@bb-sensors.com | bb-sensors.com 9 / 24
USING THE LOGGER
When the logger and PC are connected, the Configurator is fully loaded. Four screens are available: Communication, Channels,
Acquisitions and Diagnosis; which can be selected in the upper tabs.
CHANNELS SCREEN
In the channels selection screen, the user defines features for each channel, such as: input type, alarms, scale etc. Functions
displayed are:
Figure 2 – Channels screen
Title – Defines a name for the logger (up to 16 characters).
Channel selection – Selects the channel to be configured. Information presented/defined in the remaining fields for this screen are
related to the channel selected. Define channels that will remain active or enabled while the logger is operating in this field. It is also
possible to assign names to the input channels (up to 8 characters).
Input Type – Defines the input type for the channel selected in Channel Selection. Options are displayed in Table 01.
Measurement Unit – Defines the measurement unit to be used. For temperature sensor input, options are ºC or ºF. For 4-20 mA
and 0- 50 mV inputs it is possible to write the required unit.
Scale – For 4-20 m A and 0-50 mV input types it makes possible to make an indication scale, with the upper and lower indication
limits. For the other input types, the logger use a preset scale.
Decimal – It defines the number of decimal points to be displayed in scales configured for channels with 4-20 mA and 0-50 mV
inputs.
Decimal point – When the channel is configured as thermocouple, it is possible to choose how values will be displayed: with or
without the decimal point.
ALM1/ALM2 – It makes possible to define if ALM1 and ALM2 will be used as Alarm or Digital Output. As Alarm, it is triggered as
defined
in Alarm 1 and Alarm 2. In Digital Output, triggering takes place only through serial communication commands (Preset Single Coil).
Alarm 1/Alarm 2 – After the alar m outputs are defined, the user defines alarm Set Points in these fields; as well as the working
mode as Low or High. It also defines which output will be connected to the alarm.
The association between an output and alarms 1 and 2 is not required, but the alarm activities will be indicated in the Holding
Register 0008, and will be available to be read through serial communication.

INSTRUCTION MANUAL
FieldLogger
with 128k Memory
0141 0315-57 01. April 2009
B+B Thermo-Technik Gmb H | Hei nrich-Hertz-Str. 4 | D-78166 Donaueschingen
Technical changes reserved Fon +49 771 83160 | Fax +49 771 831650 | info@bb-sensors.com | bb-sensors.com 10 / 24
ACQUISITIONS SCREEN
Parameters related to how and when readings take place are defined in this screen. Acquisition is the operation of reading and
registering in local memory the process variables (temperature, pressure etc) present in the input channels enabled. Among
parameters are acquisitions start and end of readings and time interval between acquisitions.
In models without local memory, aimed only at measuring and making them available in the supervisory software, this screen will not
be enabled. The acquisition screen parameters are:
Figure 3 – Acquisitions screen
Acquisitions start – There are 3 options to start acquisitions in thelogger:
1. Specific date and time. Day, month, year, hour, minutes and seconds are defined. When they are reached they start reading
signals from the input channels of the logger.
2. Digital input (DI) activation. In this mode, acquisitions start whenever the DI is triggered and stop when it is disabled.
3. Through Start button – Determines that acquisitions start when the Start button is pushed from the screen. In this option,
observe that settings must be already configured and applied to the logger.
Base Interval Between Loggings – Defines the minimum time interval between acquisitions. It can be from hours up to tenths of
seconds and it is common to all channels. However, a multiplier of the base interval can be defined for each active channel.
Interval multiplier – Readings can be taken in different times for each channel, a base interval multiplier is defined for each
channel. The multiplier can be defined between 1 and 255.
Example: For a base interval of 10s, the logger will record at every 10s the value measured in channels programmed with multiplier
equals 1. At every 20s in channels with multiplier equals 2,…
Acquisition end – Defines the acquisitions end. Options are:
1. Specific date and time. Day, month, year, hour, minutes and seconds are defined. When they are reached they stop acquiring
values from the input channels of the logger.
Note: Acquisitions can be stopped before the preset period is reached by pressing the Stop Now button or when the full
memory capacity is reached.
2. Stop after a defined number of base intervals is reached.
Note: Acquisitions can be stopped before the expected acquisitions number is reached by pressing the Stop Now button or
when the full memory capacity is reached.
3. Never stop (wrap around). Acquisitions are taken continuously. When the full memory capacity is reached, older information
are replaced by new ones. The process is only interrupted if the Stop Now button is pressed.

INSTRUCTION MANUAL
FieldLogger
with 128k Memory
0141 0315-57 01. April 2009
B+B Thermo-Technik Gmb H | Hei nrich-Hertz-Str. 4 | D-78166 Donaueschingen
Technical changes reserved Fon +49 771 83160 | Fax +49 771 831650 | info@bb-sensors.com | bb-sensors.com 11 / 24
4. Full memory. Acquisitions end when the full memory capacity is reached. Pressing the Stop Now button also ends
acquisitions.
The Acquisitions screen has also an important group of information about the logger and acquisitions:
Current Time: Displays current date and time used by the computer.
Logger Time: Displays current date and time used by the logger.
Logging Available Memory: Displays the total memory capacity for the logger.
Programmed Number of Acquisitions: Indicates the number of acquisitions that will be performed and logged in memory. It is also
possible to check the number of acquisitions per channel enabled.
Programmed Number of Base Intervals: Informs the minimum interval between programmed acquisitions
Estimated time: Informs time required to complete the programmed acquisitions.
After setting up the logger and estimating possible implications, the user must send the configuration to the logger. Press Apply.
To send and leave the configurator software, press OK. To leave the configuration without sending to the logger, press Cancel.
Note: When the new configuration is sent, the logger time is updated, adopting the computer time (date and hour).
COMMUNICATION SCREEN
This screen is showed even when there is no communication between logger and computer. It provides parameters related to
system communication. Its fields are:
Host Parameters (HP) – Provides a choice for the serial communication port to be used.
Network Operation – Makes possible to define a new communication address and baud rate for the logger. Define values and press
Change. If connected, it makes possible to read the complete configuration of a given logger. It is only necessary to select the
required address and press Read Configurations. The configurator starts to display the selectedlogger configuration on the screen.
Parity – Makes possible to define parity used in serial communication. The number of Stop Bits as recommended by the Modbus
standard depends on parity. When parity is configured as none, 2 stop bits are used; when parity is configured as Odd or Even, 1
stop bit is used.
Note: Baud rate and parity must be the same for all the network elements.
Automatic Search – Makes possible to search the connected logger communication parameters, in a point-to-point communication,
testing all possibilities of addresses and possible transmission rates (baud rate). After the parameters are found, they are displayed.
Figure 4 – Communication screen

INSTRUCTION MANUAL
FieldLogger
with 128k Memory
0141 0315-57 01. April 2009
B+B Thermo-Technik Gmb H | Hei nrich-Hertz-Str. 4 | D-78166 Donaueschingen
Technical changes reserved Fon +49 771 83160 | Fax +49 771 831650 | info@bb-sensors.com | bb-sensors.com 12 / 24
DIAGNOSTIC PAGE
In this screen, the user visualizes how measured variables behave. Channels are presented with values measured (updated at every
second), names, units, status and alarms triggered.
Information about the logger such as serial number, version and operation status are also displayed.
It is possible to apply digital filters to input channels using this screen. These filters are intended to reduce oscillations in readings.
Intensity can be adjusted from 0 to 9 and the factory set is level 3. It is also possible to display the local power supply frequency so
that derivative oscillations are minimized.
Figure 5 – Diagnostic screen

INSTRUCTION MANUAL
FieldLogger
with 128k Memory
0141 0315-57 01. April 2009
B+B Thermo-Technik Gmb H | Hei nrich-Hertz-Str. 4 | D-78166 Donaueschingen
Technical changes reserved Fon +49 771 83160 | Fax +49 771 831650 | info@bb-sensors.com | bb-sensors.com 13 / 24
9.5 INSTALLATION
The logger has two terminal lines for input connections, power supply, backup power supply (VBAT), digital input and serial
communication.
INSTALLATION RECOMMENDATIONS
• Input signal leads must be installed in grounded conduits and away from power or contactor wires.
• Instruments must be powered only by an exclusive power supply.
• Installing RC filters (47W and 100nF, serial) is strongly recommended at contactor coils or any other inductors.
• System failure should always be taken into account when designing a control panel to avoid irreversible damage to equipment or
people. The ALM1 and ALM2 output relays, used as alarms, do not warrant total protection.
• Wires section: minimum gauge 0.14 mm².
Electrical Connections – Inputs
Figure 8 – Current (4 to 20 mA) Figure 9 – Voltage (0 to 50 mV)

INSTRUCTION MANUAL
FieldLogger
with 128k Memory
0141 0315-57 01. April 2009
B+B Thermo-Technik Gmb H | Hei nrich-Hertz-Str. 4 | D-78166 Donaueschingen
Technical changes reserved Fon +49 771 83160 | Fax +49 771 831650 | info@bb-sensors.com | bb-sensors.com 14 / 24
ELECTRICAL CONNECTIONS – POWER SUPPLY
Terminals 19 and 20 identified with POWER show the main power supply of the logger. As a backup power supply, in VBAT,
terminals 14 and 15, an extra power supply of +6 to +24 VDC can be connected so that the device can continue measuring and
logging operations in the failure of a main power supply.
While using the VBAT supply the logger can not use the serial communication and does not trigger ALM1 and ALM2
outputs.
ELECTRICAL CONNECTIONS – ALARMS AND DIGITAL INPUT
ALM1 and ALM2 outputs, terminals 21, 22, 23 and 24, correspond to relays used by alarms. NA contacts are available, with load
capacity of 3 A / 250 V.
A dry contact device should be used (switch, relay etc) in digital inputs, terminals 13 and 14. When the contact is closed, the digital
input is triggered.
Notice that when a digital input terminal is common to the VBAT negative it can be concurrently used to both ends.
9 0 1 2 3 4 5 6 7 8 9 0 1 2 3 4 5 6
Po
wer
Alar
m 1
Alar
m 2
Ch.
5
Ch.
6
Ch.
7
Ch.
8
Figure 10 - terminal lay-out

INSTRUCTION MANUAL
FieldLogger
with 128k Memory
0141 0315-57 01. April 2009
B+B Thermo-Technik Gmb H | Hei nrich-Hertz-Str. 4 | D-78166 Donaueschingen
Technical changes reserved Fon +49 771 83160 | Fax +49 771 831650 | info@bb-sensors.com | bb-sensors.com 15 / 24
10. OPERATION
After configuration and installation the logger is ready to use. When you turn on the logger, the LOGGING indicator flashes in the
front panel, this indicates that the logger is operating properly. It will flash once at every 2 seconds when it is not logging and twice at
every two seconds when logging. Acquisitions start and end according to what was defined during configuration. The speed of input
channels readings depends on the number of channels enabled and type of input programmed for each channel.
The logger starts readings immediately after it is turned on. Current values found are temporarily stored in internal registers. The
logger sends the content of its internal registers when the supervisory software requires.
In the front panel there are two indicators that show when the logger is communicating with the external world. When the Rx
indicator is turned on, it means that the device is receiving data; the Tx indicator shows that data ar e sent outside through serial
interface.
ACQUIRED DATA COLLECTION
For the collection of data logged in local memory, it is necessary to use FieldChart or Superview software, or to build an application
in a supervisory software whose version supports the specific driver.
10.1 SERIAL COMMUNICATION
The logger has a master-slave asynchronous serial communication interface RS485 to communicate with a supervisory computer
(master). The logger is always the slave.
Communication always start with the master, which sends a command to the slave address with whom it wants to communicate. The
addressed slave undertakes the command and sends the corresponding answer to the master.
The logger also accepts broadcast commands.
Note: Recommended cables for serial communication are 3 x 24 AWG (3 x 0,25mm
2
)+ shield. This way the shield can be grounded
while cables are used for signals.

INSTRUCTION MANUAL
FieldLogger
with 128k Memory
0141 0315-57 01. April 2009
B+B Thermo-Technik Gmb H | Hei nrich-Hertz-Str. 4 | D-78166 Donaueschingen
Technical changes reserved Fon +49 771 83160 | Fax +49 771 831650 | info@bb-sensors.com | bb-sensors.com 16 / 24
FEATURES
Signals supported by RS485 standard. We recommend reading the document “Basic concepts of RS485 and RS422” available in the
CD that is delivered with the equipment.
Communication signals are electrically isolated from the rest of the device, with configurable speed of 1200, 2400, 4800, 9600,
19200 and 57600 bps.
Number of data bits: 8.
Parity: none, odd or even.
Number of stop bits: 1 (odd or even parity) or 2 (no parity).
Response delay: up to 20 ms after command is received.
Protocol: Modbus RTU, available in most supervisory software of the industry.
RS-485 signals are:
D1 D D+ B Bidirectional Data Line
D- A Reserved Bidirectional Data
line
enhances the Communication
Performance
16
GND
PHYSICAL SPECIFICATIONS
• Three-wire connection (2 data wires and 1 common) between 1 master and up to 247 instruments (maximum of 31 per network
segment) in bus topology.
• Shielded cable, 3 x 24 AWG (3 x 0,25mm
2
), with shield grounded in one end;
• 16 pF capacitance at every 30 m;
• Typical impedance: 120 Ohms;
• Maximum length 1200 m;
USB-i485

INSTRUCTION MANUAL
FieldLogger
with 128k Memory
0141 0315-57 01. April 2009
B+B Thermo-Technik Gmb H | Hei nrich-Hertz-Str. 4 | D-78166 Donaueschingen
Technical changes reserved Fon +49 771 83160 | Fax +49 771 831650 | info@bb-sensors.com | bb-sensors.com 17 / 24
10.2 REGISTERS
The logger configurable parameters are organized in a Table of Registers. The block reading command makes possible to read
multiple registers with the same request.
The configuration software, however, makes such table not necessary, once it reads and writes in all registers that require
configuration. Each parameter in the table is a word of 16 bits.
The available Modbus command for the reading operation is the following:
Current Value of Channel 1
Bytes 2 and 3 of Channel 5 String (tag)
Current Value of Channel 2
Bytes 4 and 5 of Channel 5 String (tag)
Current Value of Channel 3
Bytes 6 and 7 of Channel 5 String (tag)
Current Value of Channel 4
Bytes 0 and 1 of Channel 6 String (tag)
Current Value of Channel 5
Bytes 2 and 3 of Channel 6 String (tag)
Current Value of Channel 6
Bytes 4 and 5 of Channel 6 String (tag)
Current Value of Channel 7
Bytes 6 and 7 of Channel 6 String (tag)
Current Value of Channel 8
Bytes 0 and 1 of Channel 7 String (tag)
Bytes 2 and 3 of Channel 7 String (tag)
Reserved: Inform General
Status
0110 Bytes 4 and 5 of Channel 7 String (tag)
Informs Channel 1 and 2 Status
Bytes 6 and 7 of Channel 7 String (tag)
Informs Channel 3 and 4 Status
Bytes 0 and 1 of Channel 8 String (tag)
Informs Channel 5 and 6 Status
Bytes 2 and 3 of Channel 8 String (tag)
Informs Channel 7 and 8 Status
Bytes 4 and 5 of Channel 8 String (tag)
Current Time (seconds and
minute)
0115 Bytes 6 and 7 of Channel 8 String (tag)
0015 Current Time (hour and day) 0116
Bytes 0 and 1 of the Engineering Unit of
Channel 1
0016 Current Time (month and year) 0117
Bytes 2 and 3 of the Engineering Unit of
Channel 1
0017 Channel 1 Configuration 0118
Bytes 4 and 5 of the Engineering Unit of
Channel 1
0018 Channel 2 Configuration 0119
Bytes 6 and 7 of the Engineering Unit of
Channel 1
0019 Channel 3 Configuration 0120
Bytes 0 and 1 of the Engineering Unit of
Channel 2
0020 Channel 4 Configuration 0121
Bytes 2 and 3 of the Engineering Unit of
Channel 2
0021 Channel 5 Configuration 0122
Bytes 4 and 5 of the Engineering Unit of
Channel 2
0022 Channel 6 Configuration 0123
Bytes 6 and 7 of the Engineering Unit of
Channel 2
0023 Channel 7 Configuration 0124
Bytes 0 and 1 of the Engineering Unit of
Channel 3
0024 Channel 8 Configuration 0125
Bytes 2 and 3 of the Engineering Unit of
Channel 3
0076 Bytes 0 and 1 of the Title String 0126
Bytes 4 and 5 of the Engineering Unit of
Channel 3
0077 Bytes 2 and 3 of the Title String 0127
Bytes 6 and 7 of the Engineering Unit of
Channel 3

INSTRUCTION MANUAL
FieldLogger
with 128k Memory
0141 0315-57 01. April 2009
B+B Thermo-Technik Gmb H | Hei nrich-Hertz-Str. 4 | D-78166 Donaueschingen
Technical changes reserved Fon +49 771 83160 | Fax +49 771 831650 | info@bb-sensors.com | bb-sensors.com 18 / 24
0078 Bytes 4 and 5 of the Title String 0128
Bytes 0 and 1 of the Engineering Unit of
Channel 4
0079 Bytes 6 and 7 of the Title String 0129
Bytes 2 and 3 of the Engineering Unit of
Channel 4
0080 Bytes 8 and 9 of the Title String 0130
Bytes 4 and 5 of the Engineering Unit of
Channel 4
Bytes 10 and 11 of the Title
String
Bytes 6 and 7 of the Engineering Unit of
Channel 4
Bytes 12 and 13 of the Title
String
Bytes 0 and 1 of the Engineering Unit of
Channel 5
Bytes 14 and 15 of the Title
String
Bytes 2 and 3 of the Engineering Unit of
Channel 5
Bytes 0 and 1 of Channel 1
String (tag)
Bytes 4 and 5 of the Engineering Unit of
Channel 5
Bytes 2 and 3 of Channel 1
String (tag)
Bytes 6 and 7 of the Engineering Unit of
Channel 5
Bytes 4 and 5 of Channel 1
String (tag)
Bytes 0 and 1 of the Engineering Unit of
Channel 6
Bytes 6 and 7 of Channel 1
String (tag)
Bytes 2 and 3 of the Engineering Unit of
Channel 6
Bytes 0 and 1 of Channel 2
String (tag)
Bytes 4 and 5 of the Engineering Unit of
Channel 6
Bytes 2 and 3 of Channel 2
String (tag)
Bytes 6 and 7 of the Engineering Unit of
Channel 6
Bytes 4 and 5 of Channel 2
String (tag)
Bytes 0 and 1 of the Engineering Unit of
Channel 7
Bytes 6 and 7 of Channel 2
String (tag)
Bytes 2 and 3 of the Engineering Unit of
Channel 7
Bytes 0 and 1 of Channel 3
String (tag)
Bytes 4 and 5 of the Engineering Unit of
Channel 7
Bytes 2 and 3 of Channel 3
String (tag)
Bytes 6 and 7 of the Engineering Unit of
Channel 7
Bytes 4 and 5 of Channel 3
String (tag)
Bytes 0 and 1 of the Engineering Unit of
Channel 8
Bytes 6 and 7 of Channel 3
String (tag)
Bytes 2 and 3 of the Engineering Unit of
Channel 8
Bytes 0 and 1 of Channel 4
String (tag)
Bytes 4 and 5 of the Engineering Unit of
Channel 8
Bytes 2 and 3 of Channel 4
String (tag)
Bytes 6 and 7 of the Engineering Unit of
Channel 8
Bytes 4 and 5 of Channel 4
String (tag)
0152 Decimal Point Position for Channels
0099
Bytes 6 and 7 of Channel 4
String (tag)
0199 Software Version
0100
Bytes 0 and 1 of Channel 5
String (tag)

INSTRUCTION MANUAL
FieldLogger
with 128k Memory
0141 0315-57 01. April 2009
B+B Thermo-Technik Gmb H | Hei nrich-Hertz-Str. 4 | D-78166 Donaueschingen
Technical changes reserved Fon +49 771 83160 | Fax +49 771 831650 | info@bb-sensors.com | bb-sensors.com 19 / 24
NOTES ABOUT REGISTERS
Registers 00 to 07:
They keep the continuously updated value of readings from their respective input channels. The speed of such update depends on
the number of channels enabled and the input type selected for such channels. As the register is constantly scanning inputs, the
higher the number of active channels the higher the number of inputs that will be read until the device reads the same channel again,
decreasing the update speed. Besides, channels configured as Pt100 are read twice each, as part of a compensation algorithm of
the cable resistance. In cases where there is only one active channel (no defined as Pt100), this channel is updated at every 0.2
seconds (best rate). In the case of eight Pt100 channels (worst rate), complete scanning of channels takes 0.95 seconds.
The signal of the value read is represented as complement of 2.
In cases with resolutions in tenth of degrees, such as Pt100 and thermocouples, the value is read without the decimal point
(multiplied by 10), for protocol constraints. The same way, the decimal point is disregarded in linear inputs (4-20 mA e 0-50 mV).
Register 08:
Reports if there is an alarm event in some of the channels and if the event corresponds to a Low or High alarm.
Byte High:
Byte Low:
Registers 10 to 13
Report A/D conversion errors.
byte high -> odd channel:
byte low ® even channel:
Same as byte high for the adjacent channel.

INSTRUCTION MANUAL
FieldLogger
with 128k Memory
0141 0315-57 01. April 2009
B+B Thermo-Technik Gmb H | Hei nrich-Hertz-Str. 4 | D-78166 Donaueschingen
Technical changes reserved Fon +49 771 83160 | Fax +49 771 831650 | info@bb-sensors.com | bb-sensors.com 20 / 24
Registers 14 to 16
Only valid when FieldLogger has a real-time clock (model with memory)
byte high: Second (Reg. 14), hour (Reg. 15) or month (Reg. 16) device currencies.
byte low: Minute (Reg. 14), day (Reg. 15) or year (Reg. 16) device currencies. Year is always based on the number 1998. Thus, if
the value logged is 9, in fact the current year is (1998 + 9 =) 2007.
Registers 17 to 24:
Deal with the current configuration of their respective channels, more specifically about the type of input, engineering unit for
temperature sensors (Pt100 and thermocouple), channel acquisitions enablement, option of differential input (sensor between two
inputs) and type of alarm triggering.
byte high:
Byte low:

INSTRUCTION MANUAL
FieldLogger
with 128k Memory
0141 0315-57 01. April 2009
B+B Thermo-Technik Gmb H | Hei nrich-Hertz-Str. 4 | D-78166 Donaueschingen
Technical changes reserved Fon +49 771 83160 | Fax +49 771 831650 | info@bb-sensors.com | bb-sensors.com 21 / 24
-50.0 to 760.0°C (-58.0 to 1400.0°F)
-90.0 to 1370.0°C (-130.0 to 2498.0°F)
-100.0 to 400.0°C (-148.0 to 752.0°F)
-35.0 to 720.0°C (-31.0 to 1328.0°F)
-90.0 to 1300.0°C (-130.0 to 2372.0°F)
0.0 to 1760.0°C (-32.0 to 3200.0°F)
0.0 to 1760.0°C (-32.0 to 3200.0°F)
150.0 to 1820.0°C (302.0 to 3308.0°F)
-200.0 to 530.0°C (-328.0 to 986.0°F)
Thermocouple J curve, -50.0 to 760.0°C
Thermocouple K curve, -90.0 to 1370.0°C
Thermocouple T curve, -100.0 to 400.0°C
Thermocouple E curve, -35.0 to 720.0°C
Thermocouple N curve, -90.0 to 1300.0°C
Thermocouple R curve, 0.0 to 1760.0°C
Thermocouple S curve, 0.0 to 1760.0°C
Thermocouple B curve, 150.0 to 1820.0°C
Pt100 curve, -200.0 to 530.0°C
Programmable indication range from -1999 to 9999
Programmable indication range from -1999 to 9999
Registers 76 to 147:
Contain bytes with ASCII characters of the general title, individual tags of each channel and respective engineering units.
Register 152:
Reports the decimal point position of channels configured as linear inputs. At every two bits there is the information about a channel:
00 no decimal point
01 one decimal point
10 two decimal points
11 three decimal points
byte high:
byte high:

INSTRUCTION MANUAL
FieldLogger
with 128k Memory
0141 0315-57 01. April 2009
B+B Thermo-Technik Gmb H | Hei nrich-Hertz-Str. 4 | D-78166 Donaueschingen
Technical changes reserved Fon +49 771 83160 | Fax +49 771 831650 | info@bb-sensors.com | bb-sensors.com 22 / 24
TRIGGERING DIGITAL OUTPUTS
The Modbus command available to enable and disable outputs (provided they are configured as digital outputs) is the following:
05 – Preset Single Coil (Writes in Digital Outputs)
The ALM1 output corresponds to the output point address 0. The ALM2 output corresponds to the output point address 1.
COMMUNCATION FAULT
In case there is any communication issue with the device, please check the following items:
• Is the device correctly energized?
• Is the serial cable tightly connected? In negative case, connect it in the appropriate form.
• In Configurator, the communication port selected corresponds to the port to which the register is connected? In case it does not
match the right port, select the correct one.
• Is there any other software using the required communication port in the computer? In positive case, close the software and try
to communicate again.
If none of the above steps manage to solve the problem try to:
• Run the “Automatic Search” option, and check if there is not more than one register connected to the network.
• Use other serial port from PC.
• Install the Configurator in another PC and try to repeat the steps indicated above

INSTRUCTION MANUAL
FieldLogger
with 128k Memory
0141 0315-57 01. April 2009
B+B Thermo-Technik Gmb H | Hei nrich-Hertz-Str. 4 | D-78166 Donaueschingen
Technical changes reserved Fon +49 771 83160 | Fax +49 771 831650 | info@bb-sensors.com | bb-sensors.com 23 / 24
11. TECHNICAL DATA
Configurable according to Table 01. Eight (8) channels: CH1, CH2, CH3, CH4,
CH5, CH6, CH7 and CH8
Note: All thermocouples are linearized through software and cold junction
compensation. NBR 12771 calibrated thermocouples, NBR 13773 Pt100.
J, K, T, E, N: 0.2 % of the max. range ±1 °C
R, S and B: 0.25 % of the max. range ±3 °C;
Pt100: 0.2 % of the max. range
4-20 mA and 0-50 mV: 0.2 % of the max. range
Thermocouples/Pt100/0-50 mV: >1 MW, 4-20 mA: 33 W + 1.5 V
3-wire circuit with excitation current of 0.17mA and cable-resistance
compensation.
Note: When using Pt100 calibrators to gauge the meter, check if the required
excitation current is compatible with the register current
From 550 ms to 950 ms for eight channels
SPST-NA 3 A / 250 V (3 A / 30 VDC)
100 to 240 VAC/DC, 50/60 Hz, Optional 24 VAC/DC,
Max. Consumption: 2 VA
Secondary Supply (V
BAT
):
Voltage: from +6 to 24 V, Consumption: < 50 mA
Operating Temperature: 5 to 50 °C
Relative humidity: 80% to 31°C, linear fall of 50% to 40 °C; Internal, use;
Installation Category II, Pollution Degree 2; altitude < 2,000 m
Standard serial communication RS485 MODBUS-RTU
FieldLogger with 128k 230VAC
0568 0040
FieldLogger, 8 Channel Universal Input Data
Logger and RTU, with 128k Internal Recording Memory 230VAC power supply
FieldLogger with 128K 24VDC
0568 0040-01
FieldLogger, 8 Channel Universal Input Data
Logger and RTU, with 128k Internal Recording Memory and 24VDC power supply
USB – i458 Converter
0568 0040-10
USB Converter to Dual RS485 Isolated Ports
RS485 / 232 Dual Isolated Converter with Power Supply

INSTRUCTION MANUAL
FieldLogger
with 128k Memory
0141 0315-57 01. April 2009
B+B Thermo-Technik Gmb H | Hei nrich-Hertz-Str. 4 | D-78166 Donaueschingen
Technical changes reserved Fon +49 771 83160 | Fax +49 771 831650 | info@bb-sensors.com | bb-sensors.com 24 / 24
12. GENERAL QUESTIONS
If you have any questions concerning this product of B+B Thermo-Technik GmbH, please do not hesitate to contact us at:
B+B Thermo-Technik GmbH
Heinrich-Hertz-Straße 4
78166 Donaueschingen
Germany
Tel.: +49 771 83160
Fax: +49 771 831650
E-Mail: info@bb-sensors.com
www.bb-sensors.com
We wish you a successful measuring!
Your Temperature-Partner
B+B Thermo-Technik GmbH
All technical information in this document are proved by us with high accuracy and shall inform you about all application possibilities. This information’s are not
confirmed by us and need to be proved by ever y user in regard to his intende d use of the equi pment. All foreign trade mark rights need to be considered.